FYI there are some rather flat video filter ICs that have been made in the past. The 6th order HMC1023LP5E is tunable, at its 28MHz setting its flat then down 0.1dB in the teens, down 0.35dB at 20 MHz. That same setting is 60dB down at about 90 MHz. It is also a dual part, designed for matched I-Q filtering.

Bob LaJeunesse

Sent: Sunday, May 15, 2022 at 5:29 PM
From: "ed breya" eb@telight.com
To: time-nuts@lists.febo.com
Subject: [time-nuts] Noise down-converter project

Continuing on, the mixer's output looks amazingly good. The filter's,
not so much. I have the IF now going directly to the SA input - no pads,
no filters, no nothing, except some SMB cable/adapter stuff, and about
20 feet of BNC cable. It looks great, letting the SA do the filtering.
The low end is a beautiful down-converted replica of the 50-90 MHz noise
signal.

I can't make high precision measurements here - most are eyeball
estimates from the SA screen, but everything is in the right ballpark,
and makes sense. The amplitude measurements depend on the SA's IF RBW
setting, which is 3 MHz maximum. The measured levels agree well with
different RBW settings. The video BW also affects it some, since extra
filtering is needed sometimes to smooth the curves.

The spec of the 4647 says the effective noise BW is 48.2 MHz. The IF
passes through the -3 dB point near 24 MHz, in close agreement. The
level is very flat (no discernible deviation), to around 20 MHz, where
it just visibly starts to curve into the band edge. The maximum PSD
appears to be around -80 to -83 dBm/Hz, estimated from the displayed
levels at different RBWs.

So, the desired signal is wonderful, if only it didn't include
everything else above. What I need is a very good LPF to get the job
done - the usual problem.

The actual filter I've been using does a good job on the higher
frequencies, but is poor on flatness. It has about 2-3 dB p-p passband
ripple, with periodicity around 5-7 MHz. I've tried various padding
arrangements at both ends, all of which tend to flatten it only a little
bit at best. Looking at it with the TG/SA setup, the character is
intrinsic to filter, and not due to just its reaction to the mixer and
cabling and such.

I hate building filters. Designing them in principle is easy, with all
sorts of available tools online, but actually rounding up the real parts
(and their parasitics) and physical implementation is a PITA. But, I
suppose I'll have to do it eventually for this project. I know how nice
it can be, with the right filter, but for now, I'll have to go with what
I have.

This particular filter is a packaged module type that I've had for a
long time, and used in many experimental setups. In fact, I had to
borrow it from its commitment to another project. Despite its
limitations, it can be very handy, and it is very simple inside, so I'd
like to replicate it for other uses. I plan to open a thread about this
as a separate issue.

In the mean time, it will be for this noise project, and I'll have some
more to report, so next up will be the low frequency/DC aspects.

Ed


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

FYI there are some rather flat video filter ICs that have been made in the past. The 6th order HMC1023LP5E is tunable, at its 28MHz setting its flat then down 0.1dB in the teens, down 0.35dB at 20 MHz. That same setting is 60dB down at about 90 MHz. It is also a dual part, designed for matched I-Q filtering. Bob LaJeunesse > Sent: Sunday, May 15, 2022 at 5:29 PM > From: "ed breya" <eb@telight.com> > To: time-nuts@lists.febo.com > Subject: [time-nuts] Noise down-converter project > > Continuing on, the mixer's output looks amazingly good. The filter's, > not so much. I have the IF now going directly to the SA input - no pads, > no filters, no nothing, except some SMB cable/adapter stuff, and about > 20 feet of BNC cable. It looks great, letting the SA do the filtering. > The low end is a beautiful down-converted replica of the 50-90 MHz noise > signal. > > I can't make high precision measurements here - most are eyeball > estimates from the SA screen, but everything is in the right ballpark, > and makes sense. The amplitude measurements depend on the SA's IF RBW > setting, which is 3 MHz maximum. The measured levels agree well with > different RBW settings. The video BW also affects it some, since extra > filtering is needed sometimes to smooth the curves. > > The spec of the 4647 says the effective noise BW is 48.2 MHz. The IF > passes through the -3 dB point near 24 MHz, in close agreement. The > level is very flat (no discernible deviation), to around 20 MHz, where > it just visibly starts to curve into the band edge. The maximum PSD > appears to be around -80 to -83 dBm/Hz, estimated from the displayed > levels at different RBWs. > > So, the desired signal is wonderful, if only it didn't include > everything else above. What I need is a very good LPF to get the job > done - the usual problem. > > The actual filter I've been using does a good job on the higher > frequencies, but is poor on flatness. It has about 2-3 dB p-p passband > ripple, with periodicity around 5-7 MHz. I've tried various padding > arrangements at both ends, all of which tend to flatten it only a little > bit at best. Looking at it with the TG/SA setup, the character is > intrinsic to filter, and not due to just its reaction to the mixer and > cabling and such. > > I hate building filters. Designing them in principle is easy, with all > sorts of available tools online, but actually rounding up the real parts > (and their parasitics) and physical implementation is a PITA. But, I > suppose I'll have to do it eventually for this project. I know how nice > it can be, with the right filter, but for now, I'll have to go with what > I have. > > This particular filter is a packaged module type that I've had for a > long time, and used in many experimental setups. In fact, I had to > borrow it from its commitment to another project. Despite its > limitations, it can be very handy, and it is very simple inside, so I'd > like to replicate it for other uses. I plan to open a thread about this > as a separate issue. > > In the mean time, it will be for this noise project, and I'll have some > more to report, so next up will be the low frequency/DC aspects. > > Ed > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com >
G
ghf@hoffmann-hochfrequenz.de
Mon, May 16, 2022 3:11 PM

Am 2022-05-16 15:16, schrieb Robert LaJeunesse:

FYI there are some rather flat video filter ICs that have been made in
the past. The 6th order HMC1023LP5E is tunable, at its 28MHz setting
its flat then down 0.1dB in the teens, down 0.35dB at 20 MHz. That
same setting is 60dB down at about 90 MHz. It is also a dual part,
designed for matched I-Q filtering.

Declared dead at DigiKey.

Sent: Sunday, May 15, 2022 at 5:29 PM
From: "ed breya" eb@telight.com

The actual filter I've been using does a good job on the higher
frequencies, but is poor on flatness. It has about 2-3 dB p-p passband
ripple, with periodicity around 5-7 MHz. I've tried various padding
arrangements at both ends, all of which tend to flatten it only a
little
bit at best. Looking at it with the TG/SA setup, the character is
intrinsic to filter, and not due to just its reaction to the mixer and
cabling and such.

I hate building filters. Designing them in principle is easy, with all
sorts of available tools online, but actually rounding up the real
parts
(and their parasitics) and physical implementation is a PITA. But, I
suppose I'll have to do it eventually for this project. I know how
nice
it can be, with the right filter, but for now, I'll have to go with
what
I have.

Did you choose a Chebyscheff design to start with? These accept some
ripple
in the pass band, maybe some dB, to buy a steep rise of attenuation
above f-3dB.

best regards, Gerhard

Am 2022-05-16 15:16, schrieb Robert LaJeunesse: > FYI there are some rather flat video filter ICs that have been made in > the past. The 6th order HMC1023LP5E is tunable, at its 28MHz setting > its flat then down 0.1dB in the teens, down 0.35dB at 20 MHz. That > same setting is 60dB down at about 90 MHz. It is also a dual part, > designed for matched I-Q filtering. Declared dead at DigiKey. >> Sent: Sunday, May 15, 2022 at 5:29 PM >> From: "ed breya" <eb@telight.com> >> The actual filter I've been using does a good job on the higher >> frequencies, but is poor on flatness. It has about 2-3 dB p-p passband >> ripple, with periodicity around 5-7 MHz. I've tried various padding >> arrangements at both ends, all of which tend to flatten it only a >> little >> bit at best. Looking at it with the TG/SA setup, the character is >> intrinsic to filter, and not due to just its reaction to the mixer and >> cabling and such. >> >> I hate building filters. Designing them in principle is easy, with all >> sorts of available tools online, but actually rounding up the real >> parts >> (and their parasitics) and physical implementation is a PITA. But, I >> suppose I'll have to do it eventually for this project. I know how >> nice >> it can be, with the right filter, but for now, I'll have to go with >> what >> I have. Did you choose a Chebyscheff design to start with? These accept some ripple in the pass band, maybe some dB, to buy a steep rise of attenuation above f-3dB. best regards, Gerhard
G
ghf@hoffmann-hochfrequenz.de
Mon, May 16, 2022 3:46 PM

Sent: Sunday, May 15, 2022 at 5:29 PM
From: "ed breya" eb@telight.com
To: time-nuts@lists.febo.com
Subject: [time-nuts] Noise down-converter project

I just ran QUCS-STUDIO to design two filters.

The software has an ADS-touch but is much more friendly.
S-param-simulation, harmonic balance, nonlinear, design helpers,
interface to KiCad and Octave....  And it's free.

http://qucsstudio.de/de/start/    >

helpful stuff, including tutorials:

http://www.gunthard-kraus.de/  >

regards, Gerhard

>> Sent: Sunday, May 15, 2022 at 5:29 PM >> From: "ed breya" <eb@telight.com> >> To: time-nuts@lists.febo.com >> Subject: [time-nuts] Noise down-converter project I just ran QUCS-STUDIO to design two filters. The software has an ADS-touch but is _much_ more friendly. S-param-simulation, harmonic balance, nonlinear, design helpers, interface to KiCad and Octave.... And it's free. < http://qucsstudio.de/de/start/ > helpful stuff, including tutorials: < http://www.gunthard-kraus.de/ > regards, Gerhard
LJ
Lux, Jim
Mon, May 16, 2022 4:00 PM

On 5/16/22 8:11 AM, ghf@hoffmann-hochfrequenz.de wrote:

Am 2022-05-16 15:16, schrieb Robert LaJeunesse:

FYI there are some rather flat video filter ICs that have been made in
the past. The 6th order HMC1023LP5E is tunable, at its 28MHz setting
its flat then down 0.1dB in the teens, down 0.35dB at 20 MHz. That
same setting is 60dB down at about 90 MHz. It is also a dual part,
designed for matched I-Q filtering.

Declared dead at DigiKey.

Digikey is EOLing them - last time buy is July 31 2022  (in the US)

It's not entirely dead yet.  Mouser has them - they're marked EOL - but
you can buy them for ~$40 each

This is one of those parts from Hittite (HMC partnumber) and they tend
to do small runs, but on the other hand, if demand seems to pop up, they
may make them again.

On the other hand, watch out for "custom parts" that just happen to have
a Hittite part number.  At JPL, we had a vector modulator built by
Hittite, it got a standard part number, and I assume you could buy them
until they ran out. But I get emails every once in a while asking where
to get that part we referenced.

Sent: Sunday, May 15, 2022 at 5:29 PM
From: "ed breya" eb@telight.com

The actual filter I've been using does a good job on the higher
frequencies, but is poor on flatness. It has about 2-3 dB p-p passband
ripple, with periodicity around 5-7 MHz. I've tried various padding
arrangements at both ends, all of which tend to flatten it only a
little
bit at best. Looking at it with the TG/SA setup, the character is
intrinsic to filter, and not due to just its reaction to the mixer and
cabling and such.

I hate building filters. Designing them in principle is easy, with all
sorts of available tools online, but actually rounding up the real
parts
(and their parasitics) and physical implementation is a PITA. But, I
suppose I'll have to do it eventually for this project. I know how nice
it can be, with the right filter, but for now, I'll have to go with
what
I have.

Did you choose a Chebyscheff design to start with? These accept some
ripple
in the pass band, maybe some dB, to buy a steep rise of attenuation
above f-3dB.

I agree with Ed here, easy in the tool, not necessarily easy in real
life. One aspect of more "aggressive" designs - Chebyshev, Cauer, etc.
is that they tend to be more sensitive to component variations -
especially Cauer (Elliptical) because they depend on that carefully
placed zero to get the rejection close to cutoff.

On 5/16/22 8:11 AM, ghf@hoffmann-hochfrequenz.de wrote: > Am 2022-05-16 15:16, schrieb Robert LaJeunesse: >> FYI there are some rather flat video filter ICs that have been made in >> the past. The 6th order HMC1023LP5E is tunable, at its 28MHz setting >> its flat then down 0.1dB in the teens, down 0.35dB at 20 MHz. That >> same setting is 60dB down at about 90 MHz. It is also a dual part, >> designed for matched I-Q filtering. > > Declared dead at DigiKey. Digikey is EOLing them - last time buy is July 31 2022  (in the US) It's not entirely dead yet.  Mouser has them - they're marked EOL - but you can buy them for ~$40 each This is one of those parts from Hittite (HMC partnumber) and they tend to do small runs, but on the other hand, if demand seems to pop up, they may make them again. On the other hand, watch out for "custom parts" that just happen to have a Hittite part number.  At JPL, we had a vector modulator built by Hittite, it got a standard part number, and I assume you could buy them until they ran out. But I get emails every once in a while asking where to get that part we referenced. > > >>> Sent: Sunday, May 15, 2022 at 5:29 PM >>> From: "ed breya" <eb@telight.com> > >>> The actual filter I've been using does a good job on the higher >>> frequencies, but is poor on flatness. It has about 2-3 dB p-p passband >>> ripple, with periodicity around 5-7 MHz. I've tried various padding >>> arrangements at both ends, all of which tend to flatten it only a >>> little >>> bit at best. Looking at it with the TG/SA setup, the character is >>> intrinsic to filter, and not due to just its reaction to the mixer and >>> cabling and such. >>> >>> I hate building filters. Designing them in principle is easy, with all >>> sorts of available tools online, but actually rounding up the real >>> parts >>> (and their parasitics) and physical implementation is a PITA. But, I >>> suppose I'll have to do it eventually for this project. I know how nice >>> it can be, with the right filter, but for now, I'll have to go with >>> what >>> I have. > > Did you choose a Chebyscheff design to start with? These accept some > ripple > in the pass band, maybe some dB, to buy a steep rise of attenuation > above f-3dB. I agree with Ed here, easy in the tool, not necessarily easy in real life. One aspect of more "aggressive" designs - Chebyshev, Cauer, etc. is that they tend to be more sensitive to component variations - especially Cauer (Elliptical) because they depend on that carefully placed zero to get the rejection close to cutoff.
EB
ed breya
Wed, May 18, 2022 9:18 PM

Thanks all, for filter info. For reasons that will become evident when I
describe the LF/DC situation, I plan to use an all-passive LC LPF. I
assume I'll be needing a fairly high-order (like 9 or so) Butterworth
type response for good flatness, and enough stop-band rejection for the
higher frequency junk.

The filter that I'm using for now is an old (1970s - 80s?) K&L brand,
marked 4L52-20-0/0-100. I've never found specific info on it over all
the years, but I did just find some data on current products that look
similar:

https://klmicrowave.nyc3.digitaloceanspaces.com/products/attachments/_plk147_1_LLSeries.pdf

The descriptions look familiar enough to get some idea of what it should
do, except the passband ripple is nowhere near these specs (0.05 dB).
The modern part numbering scheme is different, but the "4" and the "20"
seem to jive, for 4-sections, and 20 MHz fc. It looks like K&L describes
the number of sections as the number of choke elements, and that's
what's in this filter. It has 4 chokes and 5 capacitors, so 9th order,
as I understand.

I picked this up in some junk long ago, and it was in bad shape -
someone had opened it up, and the cover was left hanging by a thread -
literally - a single 0-80 screw managed to keep the lid associated with
the rest. I thought the other corners were drilled out, so I just taped
it shut, and since it seemed to work, I started using it for experiments
over the years.

A couple weeks back, I began looking at it closely. I was going to
replace the original SMAs with SMBs for this project, so figured I'd
pull the guts out so I could try to ID the part values. The caps were
easy, just regular mica types with markings. The chokes are small
ferrite toroids, apparently identical cores. I counted the turns, and
found 20 on the outer pair, and 22 on the inner. I didn't want to risk
damage by removing any parts (the chokes are silicone-gooped to the
shielding), but I did manage to get ballpark in-circuit values using the
HP4276A LCZ meter at 20 kHz, so the error from the caps isn't too bad.

I also had a bit of good luck in finding that a screw was loose - one
that anchors the assembly to the floor of the machined Al box, and is
critical for grounding the circuit board. I had assumed this was not all
that great of a filter, barely keeping the stop-band 40 dB down, or that
maybe it would get better if the lid was properly attached. I also found
that there was enough intact thread left in the corner holes, that
digging up some 0-80 screws of just the right lengths fixed the lid
mounting.

So, I got my values, got my SMBs, and the filter is rebuilt almost like
new. It still leaks some at the higher frequencies, but it's now better
than 70 dB down, which is on par with the modern spec, which only shows
to that level. It is a fairly sharp cutoff filter, dropping about 80 dB
at 50 MHz

Here's the parts info:

Caps (pF labeled, unknown tolerance) 68, 150, 150, 150, 68
Chokes (uH +/- 20% possible measurement error) 2.25, 2.43, 2.5, 2.28
Choke ratio inner/outer 1.21 according to turns count

So, anyway, I know it's symmetric, supposedly 50 ohms, and 20 MHz fc.
Since then, I've been looking at filter design tools, trying to match
what's in there to any kind of "standard" filter response, and tweaking
fc and impedance too. So far, I've found nothing that's close.

The actual amplitude response looks very much like the Chebyshev example
that Gerhard posted, and the datasheet says that's what this product
line is, so it's probably in there somewhere. It's just that the part
values don't make sense.

Ed

Thanks all, for filter info. For reasons that will become evident when I describe the LF/DC situation, I plan to use an all-passive LC LPF. I assume I'll be needing a fairly high-order (like 9 or so) Butterworth type response for good flatness, and enough stop-band rejection for the higher frequency junk. The filter that I'm using for now is an old (1970s - 80s?) K&L brand, marked 4L52-20-0/0-100. I've never found specific info on it over all the years, but I did just find some data on current products that look similar: https://klmicrowave.nyc3.digitaloceanspaces.com/products/attachments/_plk147_1_LLSeries.pdf The descriptions look familiar enough to get some idea of what it should do, except the passband ripple is nowhere near these specs (0.05 dB). The modern part numbering scheme is different, but the "4" and the "20" seem to jive, for 4-sections, and 20 MHz fc. It looks like K&L describes the number of sections as the number of choke elements, and that's what's in this filter. It has 4 chokes and 5 capacitors, so 9th order, as I understand. I picked this up in some junk long ago, and it was in bad shape - someone had opened it up, and the cover was left hanging by a thread - literally - a single 0-80 screw managed to keep the lid associated with the rest. I thought the other corners were drilled out, so I just taped it shut, and since it seemed to work, I started using it for experiments over the years. A couple weeks back, I began looking at it closely. I was going to replace the original SMAs with SMBs for this project, so figured I'd pull the guts out so I could try to ID the part values. The caps were easy, just regular mica types with markings. The chokes are small ferrite toroids, apparently identical cores. I counted the turns, and found 20 on the outer pair, and 22 on the inner. I didn't want to risk damage by removing any parts (the chokes are silicone-gooped to the shielding), but I did manage to get ballpark in-circuit values using the HP4276A LCZ meter at 20 kHz, so the error from the caps isn't too bad. I also had a bit of good luck in finding that a screw was loose - one that anchors the assembly to the floor of the machined Al box, and is critical for grounding the circuit board. I had assumed this was not all that great of a filter, barely keeping the stop-band 40 dB down, or that maybe it would get better if the lid was properly attached. I also found that there was enough intact thread left in the corner holes, that digging up some 0-80 screws of just the right lengths fixed the lid mounting. So, I got my values, got my SMBs, and the filter is rebuilt almost like new. It still leaks some at the higher frequencies, but it's now better than 70 dB down, which is on par with the modern spec, which only shows to that level. It is a fairly sharp cutoff filter, dropping about 80 dB at 50 MHz Here's the parts info: Caps (pF labeled, unknown tolerance) 68, 150, 150, 150, 68 Chokes (uH +/- 20% possible measurement error) 2.25, 2.43, 2.5, 2.28 Choke ratio inner/outer 1.21 according to turns count So, anyway, I know it's symmetric, supposedly 50 ohms, and 20 MHz fc. Since then, I've been looking at filter design tools, trying to match what's in there to any kind of "standard" filter response, and tweaking fc and impedance too. So far, I've found nothing that's close. The actual amplitude response looks very much like the Chebyshev example that Gerhard posted, and the datasheet says that's what this product line is, so it's probably in there somewhere. It's just that the part values don't make sense. Ed
EB
ed breya
Tue, May 24, 2022 1:39 AM

I managed to build a filter, using the values for a 9th order
Butterworth, 50 ohms, 25 MHz fc. The caps were fairly straightforward to
get nearly right on in values, with one or two (paralleled) selected
micas for each spot. The chokes were tricky. I decided to use IF-can
style adjustable ones, since I managed to scrounge up a few that were
close enough. The whole thing is built on special double ground plane
board (with 0.15" via grid shorting the sides together) stock, which
took a lot of hand crafting to mount the cans and lay everything out right.

I checked it with the TG, and it looks like a filter, kind of as
expected. After much tweaking of the chokes, I got it to look fairly
good, but it's all open-loop, part-wise - the chokes are set for
appearance of the response, not necessarily right values. So, it's some
kind of LPF, but that's about all I can say. The chokes are the weak
link, since they're hard to measure accurately.

I put the filter into the noise project, and the result looks pretty
good. Measuring the actual noise output on the SA, and zooming in, I
found it was flat to less than half a dB p-p, looking at 1 dB/div. Not
bad considering my eyeball-controlled adjustment using 10 dB/div and the
TG beforehand. This flatness is the net effect of the noise output
itself, and the filter, and a little bit the SA, so pretty decent. The
fc is around 22 MHz at the "best appearance" setting, and the Z-match
seems OK. There's no pad at the filter input, and about 3 dB at the
output, then that same 20 feet of cable to the SA. The high frequency
rejection looks pretty good too, with the 70 MHz and 140 MHz (the worst
offenders) below -85 dBm. This can be improved with more grounding
enhancement, and possibly adding shielding - it's kind of open
construction now, just on the board. The chokes are fairly well
contained and shielded in the cans, but the caps are exposed.

Anyway, for this purpose, it's way better than the original filter,
which can now be returned to its other project. I'm fairly happy with it
so far, but expect it to be one of those never ending projects - always
room for improvement.

Ed

I managed to build a filter, using the values for a 9th order Butterworth, 50 ohms, 25 MHz fc. The caps were fairly straightforward to get nearly right on in values, with one or two (paralleled) selected micas for each spot. The chokes were tricky. I decided to use IF-can style adjustable ones, since I managed to scrounge up a few that were close enough. The whole thing is built on special double ground plane board (with 0.15" via grid shorting the sides together) stock, which took a lot of hand crafting to mount the cans and lay everything out right. I checked it with the TG, and it looks like a filter, kind of as expected. After much tweaking of the chokes, I got it to look fairly good, but it's all open-loop, part-wise - the chokes are set for appearance of the response, not necessarily right values. So, it's some kind of LPF, but that's about all I can say. The chokes are the weak link, since they're hard to measure accurately. I put the filter into the noise project, and the result looks pretty good. Measuring the actual noise output on the SA, and zooming in, I found it was flat to less than half a dB p-p, looking at 1 dB/div. Not bad considering my eyeball-controlled adjustment using 10 dB/div and the TG beforehand. This flatness is the net effect of the noise output itself, and the filter, and a little bit the SA, so pretty decent. The fc is around 22 MHz at the "best appearance" setting, and the Z-match seems OK. There's no pad at the filter input, and about 3 dB at the output, then that same 20 feet of cable to the SA. The high frequency rejection looks pretty good too, with the 70 MHz and 140 MHz (the worst offenders) below -85 dBm. This can be improved with more grounding enhancement, and possibly adding shielding - it's kind of open construction now, just on the board. The chokes are fairly well contained and shielded in the cans, but the caps are exposed. Anyway, for this purpose, it's way better than the original filter, which can now be returned to its other project. I'm fairly happy with it so far, but expect it to be one of those never ending projects - always room for improvement. Ed
EB
ed breya
Wed, May 25, 2022 10:16 PM

Thanks Mike, for info on LCR alternatives. It's good to know of others
out there, if needed. I have an HP4276A and HP4271A. The 4276A is the
main workhorse for all part checking, since it has a wide range of LCZ,
although limited frequency coverage (100 Hz - 20 kHz). The 4271A is 1
MHz only, and good for smaller and RF parts, but very limited upper LCR
ranges. I think it works, so I can use it if needed, but would have to
check it out and build an official lead set for it. I recall working on
it a few years ago to fix some flakiness in the controls, so not 100%
sure of its present condition.

The main difficulty I've found in measuring small chokes is more of
probing/connection problem rather than instrument limitation. For most
things, I use a ground reference converter that I built for the 4276A
many years ago. It allows ground-referenced measurements, so the DUT
doesn't have to float inside the measuring bridge. The four-wire
arrangement is extended (in modified form) all the way to a small
alligator clip ground, and a probe tip, for DUT connection, so there is
some residual L in the clip and the probe tip, which causes some
variable error, especially in attaching to very small parts and leads.
When you add in the variable contact resistance too, it gets worse.
Imagine holding a small RF can (about a 1/2 inch cube) between your
fingers, with a little clip sort of hanging from one lead, and pressing
the end of the probe tip against the other lead. All the while, there's
the variable contact forces, and effects from the relative positions of
all the pieces and fingers, and the stray C from the coil to the can to
the fingers. I have pretty good dexterity, and have managed to make
these measurements holding all this stuff in one hand, while tweaking
the tuning slug with the other.

I had planned on making other accessories like another clip lead to go
in place of the probe tip, but not yet built. I also have the official
Kelvin-style lead set that came with the unit, so that's an option that
would provide much better accuracy and consistency, but the clips are
fairly large and hard to fit in tight situations, and the DUT must
float. Anyway, I can make all sorts of improvements in holding parts and
hookup, but usually I just clip and poke and try to get close enough -
especially when I have to check a lot of parts, quickly.

The other problem is that the 4276A is near its limit for getting
measurements below 1 uH, with only two digits left for nH. The 4271A
would be much better for this, with 1 nH vs 10 nH resolution.

If I get in a situation where I need to do a lot of this (if I should
get filter madness, for instance), then I'll have to improve the tools
and methods, but I'm OK for now, having slogged through it this time.

Ed

Thanks Mike, for info on LCR alternatives. It's good to know of others out there, if needed. I have an HP4276A and HP4271A. The 4276A is the main workhorse for all part checking, since it has a wide range of LCZ, although limited frequency coverage (100 Hz - 20 kHz). The 4271A is 1 MHz only, and good for smaller and RF parts, but very limited upper LCR ranges. I think it works, so I can use it if needed, but would have to check it out and build an official lead set for it. I recall working on it a few years ago to fix some flakiness in the controls, so not 100% sure of its present condition. The main difficulty I've found in measuring small chokes is more of probing/connection problem rather than instrument limitation. For most things, I use a ground reference converter that I built for the 4276A many years ago. It allows ground-referenced measurements, so the DUT doesn't have to float inside the measuring bridge. The four-wire arrangement is extended (in modified form) all the way to a small alligator clip ground, and a probe tip, for DUT connection, so there is some residual L in the clip and the probe tip, which causes some variable error, especially in attaching to very small parts and leads. When you add in the variable contact resistance too, it gets worse. Imagine holding a small RF can (about a 1/2 inch cube) between your fingers, with a little clip sort of hanging from one lead, and pressing the end of the probe tip against the other lead. All the while, there's the variable contact forces, and effects from the relative positions of all the pieces and fingers, and the stray C from the coil to the can to the fingers. I have pretty good dexterity, and have managed to make these measurements holding all this stuff in one hand, while tweaking the tuning slug with the other. I had planned on making other accessories like another clip lead to go in place of the probe tip, but not yet built. I also have the official Kelvin-style lead set that came with the unit, so that's an option that would provide much better accuracy and consistency, but the clips are fairly large and hard to fit in tight situations, and the DUT must float. Anyway, I can make all sorts of improvements in holding parts and hookup, but usually I just clip and poke and try to get close enough - especially when I have to check a lot of parts, quickly. The other problem is that the 4276A is near its limit for getting measurements below 1 uH, with only two digits left for nH. The 4271A would be much better for this, with 1 nH vs 10 nH resolution. If I get in a situation where I need to do a lot of this (if I should get filter madness, for instance), then I'll have to improve the tools and methods, but I'm OK for now, having slogged through it this time. Ed
EB
ed breya
Fri, May 27, 2022 2:15 AM

Now that I have the "official" filter in place, I can wrap up the LF/DC
issues. This is the other extreme, so no SA here, just time domain view
with a Tek 7A22 vertical, which gets down to 10 uV/div, and has settable
BW steps from 100 Hz to 1 MHz. For very low f and DC, I use a HP3456A.
There are some limits, especially in the 7A22, which is a little flaky,
but mostly puts on a good show. In either instrument, there may be
errors caused by the large HF part of the noise up to the 25 MHz or so,
way beyond what they're trying to see.

One thing that immediately showed up is the mixer DC offset (about -1.2
mV) due mostly to imperfections in the mixer, distortion in the LO, and
LO leakage into where it doesn't belong. I built a photo-voltaic
circuit to generate a current to cancel it out, but had to wait until
other issues were settled before final design adjustments.

Why a PV generator? This relates to the fundamental design plan. You may
recall that in the earlier talk on the mixer, I wanted to be able to
have galvanic isolation of the IF port, in order to eliminate or reduce
ground loop interference. Indeed, I found out right away that this was
the way to go. On the 7A22, I could see several mV of line-related junk,
and figured it was time to lift the IF off ground. For the RF
experimenting, I had the IF chassis-grounded, but had all the provisions
in place to float the whole works, from the IF port all the way to the
front panel BNC. I chose to overdo the capacitance from the IF common to
earth, with two 100 nF caps. The common-mode chassis noise disappeared,
as expected.

But, all this forces various compromises between the requirements.
First, there's not much point in making a thing that can go essentially
all the way down to DC, and possibly at very tiny signal levels
(depending on BW and noise power level), if you can't convey the signal
to an external piece of gear or experiment without ground loop
interference. So, this isolation is necessary - it raises the
common-mode impedance of the source so that the (hopefully) small
inter-chassis voltages can't push much current between equipment.

But, this is all frequency dependent too. If the ground loop
interference has higher frequency content (like in something with a SMPS
that's not very clean), the caps isolating the floating section present
much lower CM impedance, allowing more current. For this, you'd want
minimal CM capacitance.

But, minimal CM capacitance is minimally effective in shorting out the
LO and RF at the mixer - whatever leaks through due to the limited
isolation of the mixer becomes CM and additional IF signal at the IF
port. For this, you'd want as high a CM capacitance as possible, or
solid ground (which is the non-isolated form).

So, it all boils down to making appropriate trade-offs in that CM
capacitance. As mentioned earlier, I started with 200 nF, which was
sufficient for line/harmonic interference rejection, and was a good RF
short at the mixer. Next, I tried a lower extreme of 2 nF total, which
would have been great for medium frequency rejection, but alas, not a
good enough short for the LO and RF, indicated by increasing power at
the output, and increasing DC offset - it nearly doubled it.

The present compromise is about 9 nF total (the previous 2 nF plus three
2200 pF tacked on). This seems to be pretty good, with reasonably small
(maybe -90 dBm) LO showing, and only slightly higher offset compared to
the 200 nF version. I think when all's said and done, I'll end up with
about a 10-20 nF compromise value.

There's also some CM choking involved. The most important one isolates
the LO and RF CM right at the IF port, formed with three loops (about 10
uH) through a ferrite toroid of the SMB pigtail cable the goes to the
filter. A second one will be included on the output cable to the front
panel, to help at the medium to high frequencies.

I edited the box's board ground plane to form the isolated section that
carries the filter, padding, associated interconnects, and PV generator
circuits. Since this all floats, the PV method is used, and no power
supplies or chassis ground returns (which would spoil it) are needed.
The generator is two paralleled 4N37 opto-isolators operating in PV
mode, with variable LED drive for setting the offset current.

The concept of "floating" is somewhat arbitrary. In reality, the whole
output could float to any applied voltage until something breaks down,
but I decided it was safest to just hard-clamp it to chassis ground with
Si rectifiers (1N5401). Unfortunately, their zero-bias capacitance adds
to the total CM capacitance, while they can't help with any RF shorting
at the mixer - they're too big to fit near there, and are too far
removed from the action by distance and the CM choke.

Next up will be more details. It's getting close to the end. I can tell
that it's near time to wrap up or quit this project, because the
connectors are starting to wear out from all the puts and takes of the
box into the instrument - I'd say it's well over a hundred times already.

Ed

Now that I have the "official" filter in place, I can wrap up the LF/DC issues. This is the other extreme, so no SA here, just time domain view with a Tek 7A22 vertical, which gets down to 10 uV/div, and has settable BW steps from 100 Hz to 1 MHz. For very low f and DC, I use a HP3456A. There are some limits, especially in the 7A22, which is a little flaky, but mostly puts on a good show. In either instrument, there may be errors caused by the large HF part of the noise up to the 25 MHz or so, way beyond what they're trying to see. One thing that immediately showed up is the mixer DC offset (about -1.2 mV) due mostly to imperfections in the mixer, distortion in the LO, and LO leakage into where it doesn't belong. I built a photo-voltaic circuit to generate a current to cancel it out, but had to wait until other issues were settled before final design adjustments. Why a PV generator? This relates to the fundamental design plan. You may recall that in the earlier talk on the mixer, I wanted to be able to have galvanic isolation of the IF port, in order to eliminate or reduce ground loop interference. Indeed, I found out right away that this was the way to go. On the 7A22, I could see several mV of line-related junk, and figured it was time to lift the IF off ground. For the RF experimenting, I had the IF chassis-grounded, but had all the provisions in place to float the whole works, from the IF port all the way to the front panel BNC. I chose to overdo the capacitance from the IF common to earth, with two 100 nF caps. The common-mode chassis noise disappeared, as expected. But, all this forces various compromises between the requirements. First, there's not much point in making a thing that can go essentially all the way down to DC, and possibly at very tiny signal levels (depending on BW and noise power level), if you can't convey the signal to an external piece of gear or experiment without ground loop interference. So, this isolation is necessary - it raises the common-mode impedance of the source so that the (hopefully) small inter-chassis voltages can't push much current between equipment. But, this is all frequency dependent too. If the ground loop interference has higher frequency content (like in something with a SMPS that's not very clean), the caps isolating the floating section present much lower CM impedance, allowing more current. For this, you'd want minimal CM capacitance. But, minimal CM capacitance is minimally effective in shorting out the LO and RF at the mixer - whatever leaks through due to the limited isolation of the mixer becomes CM and additional IF signal at the IF port. For this, you'd want as high a CM capacitance as possible, or solid ground (which is the non-isolated form). So, it all boils down to making appropriate trade-offs in that CM capacitance. As mentioned earlier, I started with 200 nF, which was sufficient for line/harmonic interference rejection, and was a good RF short at the mixer. Next, I tried a lower extreme of 2 nF total, which would have been great for medium frequency rejection, but alas, not a good enough short for the LO and RF, indicated by increasing power at the output, and increasing DC offset - it nearly doubled it. The present compromise is about 9 nF total (the previous 2 nF plus three 2200 pF tacked on). This seems to be pretty good, with reasonably small (maybe -90 dBm) LO showing, and only slightly higher offset compared to the 200 nF version. I think when all's said and done, I'll end up with about a 10-20 nF compromise value. There's also some CM choking involved. The most important one isolates the LO and RF CM right at the IF port, formed with three loops (about 10 uH) through a ferrite toroid of the SMB pigtail cable the goes to the filter. A second one will be included on the output cable to the front panel, to help at the medium to high frequencies. I edited the box's board ground plane to form the isolated section that carries the filter, padding, associated interconnects, and PV generator circuits. Since this all floats, the PV method is used, and no power supplies or chassis ground returns (which would spoil it) are needed. The generator is two paralleled 4N37 opto-isolators operating in PV mode, with variable LED drive for setting the offset current. The concept of "floating" is somewhat arbitrary. In reality, the whole output could float to any applied voltage until something breaks down, but I decided it was safest to just hard-clamp it to chassis ground with Si rectifiers (1N5401). Unfortunately, their zero-bias capacitance adds to the total CM capacitance, while they can't help with any RF shorting at the mixer - they're too big to fit near there, and are too far removed from the action by distance and the CM choke. Next up will be more details. It's getting close to the end. I can tell that it's near time to wrap up or quit this project, because the connectors are starting to wear out from all the puts and takes of the box into the instrument - I'd say it's well over a hundred times already. Ed
EB
ed breya
Sat, Jun 4, 2022 12:32 AM

I've been working on final design cleanup, mainly in the RF. I found
quite a bit of spurious LO harmonic content up to almost 2 GHz, with
some quite strong (-75 dBm). It was time to clean up the experimental
wiring layout, so I simplified the cabling and consolidated the RF stuff
onto the LPF board. This improved things a bit, but some spurs were
still pretty big. I presumed most of it was going right through or
around the LPF, and some due to common-mode and cavity resonances inside
the box, which can have many modes.

I added a small LPF about 300 MHz (10 pF/50 nH/10 pF), inside its own
tiny shield box, forming the last bastion of filtering, right at the
inlet of the pigtail cable that goes to the isolated SMA bulkhead
fitting, and including another CM choke (only 1 pass of cable). This
filter is high enough up (over ten times the fc of the main LPF) that
they shouldn't interact very much - they are isolated only by the 3 dB
pad in between.

All along, I've wondered what to do about the reflected power from the
main LPF, that mostly has to go back to the mixer. They are separated by
maybe 300 pSec of cable, which could be in the range for resonances at
the upper end. But, various experiments during development, including
padding the LPF input, and even making a diplexer with a 50 MHz HPF to
take the HF content into a terminator, showed no difference in the noise
output flatness, although the spurious levels likely would have changed
a little - some up, some down. So, I decided to keep it simple and just
let 'er rip, with nothing extra at the LPF input.

Things are now at levels where the fine (and subtle) details show,
mostly cable dress, and grounding. I'll probably be adding bits of
shielding here and there, and maybe fooling with some RF absorbing foam
to see if any box resonances are a problem.

Speaking of subtle effects, here's something interesting. The little
shield box for the 300 MHz LPF is a type with a fold-down lid, on a
hinge formed by thinning the sheet steel. It's only good for a few open
and close operations before the hinge breaks apart, so I kept it open
while building and testing the filter. It looked great, and the time
came to close everything up and look at the spurs again. I closed the
lid, and bent the retainer tangs a little, for good closure. Virtually
all the higher frequency spurs got a few dB worse. So, was it that the
lid isn't really grounded thoroughly, and acting as an antenna to bypass
the filter, or did it affect the choke Q or part values enough, or is it
that I also changed the cable dress a bit while putting it all back
together? I'll have to figure it out.

Anyway, it's looking pretty good right now. With everything closed up,
including the box lids, as it would be when completed, all the spurs
show around -90 dBm or less. There were maybe two dozen noticeable spurs
identified earlier. Some are now in the noise floor (around -105 dBm,
but some remain, sticking out. I think most will disappear if I figure
out that 300 MHz filter box lid, which would leave the 70 MHz as the
main offender. This isn't surprising, since it's the biggest signal of
all, and it's not filtered all that much - it's too close to the main
LPF fc, and below the 300 MHz LPF. I should be able to knock it down
enough with detail work mentioned above, and I'm also pondering ways to
make a 70 MHz trap, if it won't go away. I have a couple of 70 MHz
crystals, so I could try this kind fairly easily. Does anyone have any
handy design info for crystal notch filters in this frequency range? For
an LC trap, it looks like a single L and C would be enough to get the
job done, without interacting too much with the other filters.

Ed

I've been working on final design cleanup, mainly in the RF. I found quite a bit of spurious LO harmonic content up to almost 2 GHz, with some quite strong (-75 dBm). It was time to clean up the experimental wiring layout, so I simplified the cabling and consolidated the RF stuff onto the LPF board. This improved things a bit, but some spurs were still pretty big. I presumed most of it was going right through or around the LPF, and some due to common-mode and cavity resonances inside the box, which can have many modes. I added a small LPF about 300 MHz (10 pF/50 nH/10 pF), inside its own tiny shield box, forming the last bastion of filtering, right at the inlet of the pigtail cable that goes to the isolated SMA bulkhead fitting, and including another CM choke (only 1 pass of cable). This filter is high enough up (over ten times the fc of the main LPF) that they shouldn't interact very much - they are isolated only by the 3 dB pad in between. All along, I've wondered what to do about the reflected power from the main LPF, that mostly has to go back to the mixer. They are separated by maybe 300 pSec of cable, which could be in the range for resonances at the upper end. But, various experiments during development, including padding the LPF input, and even making a diplexer with a 50 MHz HPF to take the HF content into a terminator, showed no difference in the noise output flatness, although the spurious levels likely would have changed a little - some up, some down. So, I decided to keep it simple and just let 'er rip, with nothing extra at the LPF input. Things are now at levels where the fine (and subtle) details show, mostly cable dress, and grounding. I'll probably be adding bits of shielding here and there, and maybe fooling with some RF absorbing foam to see if any box resonances are a problem. Speaking of subtle effects, here's something interesting. The little shield box for the 300 MHz LPF is a type with a fold-down lid, on a hinge formed by thinning the sheet steel. It's only good for a few open and close operations before the hinge breaks apart, so I kept it open while building and testing the filter. It looked great, and the time came to close everything up and look at the spurs again. I closed the lid, and bent the retainer tangs a little, for good closure. Virtually all the higher frequency spurs got a few dB worse. So, was it that the lid isn't really grounded thoroughly, and acting as an antenna to bypass the filter, or did it affect the choke Q or part values enough, or is it that I also changed the cable dress a bit while putting it all back together? I'll have to figure it out. Anyway, it's looking pretty good right now. With everything closed up, including the box lids, as it would be when completed, all the spurs show around -90 dBm or less. There were maybe two dozen noticeable spurs identified earlier. Some are now in the noise floor (around -105 dBm, but some remain, sticking out. I think most will disappear if I figure out that 300 MHz filter box lid, which would leave the 70 MHz as the main offender. This isn't surprising, since it's the biggest signal of all, and it's not filtered all that much - it's too close to the main LPF fc, and below the 300 MHz LPF. I should be able to knock it down enough with detail work mentioned above, and I'm also pondering ways to make a 70 MHz trap, if it won't go away. I have a couple of 70 MHz crystals, so I could try this kind fairly easily. Does anyone have any handy design info for crystal notch filters in this frequency range? For an LC trap, it looks like a single L and C would be enough to get the job done, without interacting too much with the other filters. Ed
A
Askild
Sat, Jun 4, 2022 1:42 PM

Hi Ed,

One thing I would test, that might not help, but should be easy to test, is
to put some RF-absorber in the lid of the small shielded filter box.

Regards,
Askild

On Sat, Jun 4, 2022 at 2:37 AM ed breya via time-nuts <
time-nuts@lists.febo.com> wrote:

I've been working on final design cleanup, mainly in the RF. I found
quite a bit of spurious LO harmonic content up to almost 2 GHz, with
some quite strong (-75 dBm). It was time to clean up the experimental
wiring layout, so I simplified the cabling and consolidated the RF stuff
onto the LPF board. This improved things a bit, but some spurs were
still pretty big. I presumed most of it was going right through or
around the LPF, and some due to common-mode and cavity resonances inside
the box, which can have many modes.

I added a small LPF about 300 MHz (10 pF/50 nH/10 pF), inside its own
tiny shield box, forming the last bastion of filtering, right at the
inlet of the pigtail cable that goes to the isolated SMA bulkhead
fitting, and including another CM choke (only 1 pass of cable). This
filter is high enough up (over ten times the fc of the main LPF) that
they shouldn't interact very much - they are isolated only by the 3 dB
pad in between.

All along, I've wondered what to do about the reflected power from the
main LPF, that mostly has to go back to the mixer. They are separated by
maybe 300 pSec of cable, which could be in the range for resonances at
the upper end. But, various experiments during development, including
padding the LPF input, and even making a diplexer with a 50 MHz HPF to
take the HF content into a terminator, showed no difference in the noise
output flatness, although the spurious levels likely would have changed
a little - some up, some down. So, I decided to keep it simple and just
let 'er rip, with nothing extra at the LPF input.

Things are now at levels where the fine (and subtle) details show,
mostly cable dress, and grounding. I'll probably be adding bits of
shielding here and there, and maybe fooling with some RF absorbing foam
to see if any box resonances are a problem.

Speaking of subtle effects, here's something interesting. The little
shield box for the 300 MHz LPF is a type with a fold-down lid, on a
hinge formed by thinning the sheet steel. It's only good for a few open
and close operations before the hinge breaks apart, so I kept it open
while building and testing the filter. It looked great, and the time
came to close everything up and look at the spurs again. I closed the
lid, and bent the retainer tangs a little, for good closure. Virtually
all the higher frequency spurs got a few dB worse. So, was it that the
lid isn't really grounded thoroughly, and acting as an antenna to bypass
the filter, or did it affect the choke Q or part values enough, or is it
that I also changed the cable dress a bit while putting it all back
together? I'll have to figure it out.

Anyway, it's looking pretty good right now. With everything closed up,
including the box lids, as it would be when completed, all the spurs
show around -90 dBm or less. There were maybe two dozen noticeable spurs
identified earlier. Some are now in the noise floor (around -105 dBm,
but some remain, sticking out. I think most will disappear if I figure
out that 300 MHz filter box lid, which would leave the 70 MHz as the
main offender. This isn't surprising, since it's the biggest signal of
all, and it's not filtered all that much - it's too close to the main
LPF fc, and below the 300 MHz LPF. I should be able to knock it down
enough with detail work mentioned above, and I'm also pondering ways to
make a 70 MHz trap, if it won't go away. I have a couple of 70 MHz
crystals, so I could try this kind fairly easily. Does anyone have any
handy design info for crystal notch filters in this frequency range? For
an LC trap, it looks like a single L and C would be enough to get the
job done, without interacting too much with the other filters.

Ed


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

Hi Ed, One thing I would test, that might not help, but should be easy to test, is to put some RF-absorber in the lid of the small shielded filter box. Regards, Askild On Sat, Jun 4, 2022 at 2:37 AM ed breya via time-nuts < time-nuts@lists.febo.com> wrote: > I've been working on final design cleanup, mainly in the RF. I found > quite a bit of spurious LO harmonic content up to almost 2 GHz, with > some quite strong (-75 dBm). It was time to clean up the experimental > wiring layout, so I simplified the cabling and consolidated the RF stuff > onto the LPF board. This improved things a bit, but some spurs were > still pretty big. I presumed most of it was going right through or > around the LPF, and some due to common-mode and cavity resonances inside > the box, which can have many modes. > > I added a small LPF about 300 MHz (10 pF/50 nH/10 pF), inside its own > tiny shield box, forming the last bastion of filtering, right at the > inlet of the pigtail cable that goes to the isolated SMA bulkhead > fitting, and including another CM choke (only 1 pass of cable). This > filter is high enough up (over ten times the fc of the main LPF) that > they shouldn't interact very much - they are isolated only by the 3 dB > pad in between. > > All along, I've wondered what to do about the reflected power from the > main LPF, that mostly has to go back to the mixer. They are separated by > maybe 300 pSec of cable, which could be in the range for resonances at > the upper end. But, various experiments during development, including > padding the LPF input, and even making a diplexer with a 50 MHz HPF to > take the HF content into a terminator, showed no difference in the noise > output flatness, although the spurious levels likely would have changed > a little - some up, some down. So, I decided to keep it simple and just > let 'er rip, with nothing extra at the LPF input. > > Things are now at levels where the fine (and subtle) details show, > mostly cable dress, and grounding. I'll probably be adding bits of > shielding here and there, and maybe fooling with some RF absorbing foam > to see if any box resonances are a problem. > > Speaking of subtle effects, here's something interesting. The little > shield box for the 300 MHz LPF is a type with a fold-down lid, on a > hinge formed by thinning the sheet steel. It's only good for a few open > and close operations before the hinge breaks apart, so I kept it open > while building and testing the filter. It looked great, and the time > came to close everything up and look at the spurs again. I closed the > lid, and bent the retainer tangs a little, for good closure. Virtually > all the higher frequency spurs got a few dB worse. So, was it that the > lid isn't really grounded thoroughly, and acting as an antenna to bypass > the filter, or did it affect the choke Q or part values enough, or is it > that I also changed the cable dress a bit while putting it all back > together? I'll have to figure it out. > > Anyway, it's looking pretty good right now. With everything closed up, > including the box lids, as it would be when completed, all the spurs > show around -90 dBm or less. There were maybe two dozen noticeable spurs > identified earlier. Some are now in the noise floor (around -105 dBm, > but some remain, sticking out. I think most will disappear if I figure > out that 300 MHz filter box lid, which would leave the 70 MHz as the > main offender. This isn't surprising, since it's the biggest signal of > all, and it's not filtered all that much - it's too close to the main > LPF fc, and below the 300 MHz LPF. I should be able to knock it down > enough with detail work mentioned above, and I'm also pondering ways to > make a 70 MHz trap, if it won't go away. I have a couple of 70 MHz > crystals, so I could try this kind fairly easily. Does anyone have any > handy design info for crystal notch filters in this frequency range? For > an LC trap, it looks like a single L and C would be enough to get the > job done, without interacting too much with the other filters. > > Ed > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com >