RE
Randy Evans
Mon, Apr 21, 2014 1:51 PM
I have a Datron 1082 DMM that just recently stopped working. When I power
up, it gives a "FAIL" message on the front panel. I found a flowchart in
an abbreviated servicing manual that indicates it's from a failure of the
"Check for uncorrupted calibration store". I Assume this likely means the
cal tables are corrupted? Unfortunately, I don't have sufficient
information (e.g., schematics) to do any further troubleshooting. Can
anyone help with a complete servicing manual and/or further diagnostic help?
I did download the Datron 1071DMM O&S manual from K04BB but I don't know
how similar it is to the 1082. Can anyone comment on what the differences
are?
Thanks,
Randy Evans
randyevans2688@gmail.com
I have a Datron 1082 DMM that just recently stopped working. When I power
up, it gives a "FAIL" message on the front panel. I found a flowchart in
an abbreviated servicing manual that indicates it's from a failure of the
"Check for uncorrupted calibration store". I Assume this likely means the
cal tables are corrupted? Unfortunately, I don't have sufficient
information (e.g., schematics) to do any further troubleshooting. Can
anyone help with a complete servicing manual and/or further diagnostic help?
I did download the Datron 1071DMM O&S manual from K04BB but I don't know
how similar it is to the 1082. Can anyone comment on what the differences
are?
Thanks,
Randy Evans
randyevans2688@gmail.com
CH
Chuck Harris
Mon, Apr 21, 2014 3:03 PM
Just in case you haven't yet: Always start any investigation with
checking the power supply for voltage and ripple. Test the main
electrolytic capacitors for capacitance, and ESR.
Nothing in the unit's self testing ability can be trusted if the
power supply is putting out large amounts of ripple, or is at the
wrong voltage.
-Chuck Harris
Randy Evans wrote:
I have a Datron 1082 DMM that just recently stopped working. When I power
up, it gives a "FAIL" message on the front panel. I found a flowchart in
an abbreviated servicing manual that indicates it's from a failure of the
"Check for uncorrupted calibration store". I Assume this likely means the
cal tables are corrupted? Unfortunately, I don't have sufficient
information (e.g., schematics) to do any further troubleshooting. Can
anyone help with a complete servicing manual and/or further diagnostic help?
I did download the Datron 1071DMM O&S manual from K04BB but I don't know
how similar it is to the 1082. Can anyone comment on what the differences
are?
Thanks,
Randy Evans
Just in case you haven't yet: Always start any investigation with
checking the power supply for voltage and ripple. Test the main
electrolytic capacitors for capacitance, and ESR.
Nothing in the unit's self testing ability can be trusted if the
power supply is putting out large amounts of ripple, or is at the
wrong voltage.
-Chuck Harris
Randy Evans wrote:
> I have a Datron 1082 DMM that just recently stopped working. When I power
> up, it gives a "FAIL" message on the front panel. I found a flowchart in
> an abbreviated servicing manual that indicates it's from a failure of the
> "Check for uncorrupted calibration store". I Assume this likely means the
> cal tables are corrupted? Unfortunately, I don't have sufficient
> information (e.g., schematics) to do any further troubleshooting. Can
> anyone help with a complete servicing manual and/or further diagnostic help?
>
>
> I did download the Datron 1071DMM O&S manual from K04BB but I don't know
> how similar it is to the 1082. Can anyone comment on what the differences
> are?
>
>
>
> Thanks,
>
>
>
> Randy Evans
TK
Tom Knox
Mon, Apr 21, 2014 3:15 PM
As with many products of this age the tantalum caps are a common point of failure in the 1082.
Although this current problem sounds like a memory battery.
Thomas Knox
Date: Mon, 21 Apr 2014 11:03:56 -0400
From: cfharris@erols.com
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Datron 1082
Just in case you haven't yet: Always start any investigation with
checking the power supply for voltage and ripple. Test the main
electrolytic capacitors for capacitance, and ESR.
Nothing in the unit's self testing ability can be trusted if the
power supply is putting out large amounts of ripple, or is at the
wrong voltage.
-Chuck Harris
Randy Evans wrote:
I have a Datron 1082 DMM that just recently stopped working. When I power
up, it gives a "FAIL" message on the front panel. I found a flowchart in
an abbreviated servicing manual that indicates it's from a failure of the
"Check for uncorrupted calibration store". I Assume this likely means the
cal tables are corrupted? Unfortunately, I don't have sufficient
information (e.g., schematics) to do any further troubleshooting. Can
anyone help with a complete servicing manual and/or further diagnostic help?
I did download the Datron 1071DMM O&S manual from K04BB but I don't know
how similar it is to the 1082. Can anyone comment on what the differences
are?
Thanks,
Randy Evans
As with many products of this age the tantalum caps are a common point of failure in the 1082.
Although this current problem sounds like a memory battery.
Thomas Knox
> Date: Mon, 21 Apr 2014 11:03:56 -0400
> From: cfharris@erols.com
> To: volt-nuts@febo.com
> Subject: Re: [volt-nuts] Datron 1082
>
> Just in case you haven't yet: Always start any investigation with
> checking the power supply for voltage and ripple. Test the main
> electrolytic capacitors for capacitance, and ESR.
>
> Nothing in the unit's self testing ability can be trusted if the
> power supply is putting out large amounts of ripple, or is at the
> wrong voltage.
>
> -Chuck Harris
>
> Randy Evans wrote:
> > I have a Datron 1082 DMM that just recently stopped working. When I power
> > up, it gives a "FAIL" message on the front panel. I found a flowchart in
> > an abbreviated servicing manual that indicates it's from a failure of the
> > "Check for uncorrupted calibration store". I Assume this likely means the
> > cal tables are corrupted? Unfortunately, I don't have sufficient
> > information (e.g., schematics) to do any further troubleshooting. Can
> > anyone help with a complete servicing manual and/or further diagnostic help?
> >
> >
> > I did download the Datron 1071DMM O&S manual from K04BB but I don't know
> > how similar it is to the 1082. Can anyone comment on what the differences
> > are?
> >
> >
> >
> > Thanks,
> >
> >
> >
> > Randy Evans
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
RE
Randy Evans
Tue, Apr 22, 2014 2:23 AM
I believe the voltages are correct but I have no schematics or
troubleshooting instructions to confirm. I do see what looks like a
battery in the memory section of the 1082 and it measures 0.08 VDC. Not a
good sign. I am really hoping someone has a Datron 1082 Service manual
with schematics that they can steer me to, preferably a soft copy so I can
better troubleshoot the unit. It might just need a new battery (plus a
re-cal) to get it going again but I hate going blind.
Randy
On Mon, Apr 21, 2014 at 8:15 AM, Tom Knox actast@hotmail.com wrote:
As with many products of this age the tantalum caps are a common point of
failure in the 1082.
Although this current problem sounds like a memory battery.
Thomas Knox
Date: Mon, 21 Apr 2014 11:03:56 -0400
From: cfharris@erols.com
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Datron 1082
Just in case you haven't yet: Always start any investigation with
checking the power supply for voltage and ripple. Test the main
electrolytic capacitors for capacitance, and ESR.
Nothing in the unit's self testing ability can be trusted if the
power supply is putting out large amounts of ripple, or is at the
wrong voltage.
-Chuck Harris
Randy Evans wrote:
I have a Datron 1082 DMM that just recently stopped working. When I
up, it gives a "FAIL" message on the front panel. I found a flowchart
an abbreviated servicing manual that indicates it's from a failure of
"Check for uncorrupted calibration store". I Assume this likely means
cal tables are corrupted? Unfortunately, I don't have sufficient
information (e.g., schematics) to do any further troubleshooting. Can
anyone help with a complete servicing manual and/or further diagnostic
I did download the Datron 1071DMM O&S manual from K04BB but I don't
how similar it is to the 1082. Can anyone comment on what the
and follow the instructions there.
I believe the voltages are correct but I have no schematics or
troubleshooting instructions to confirm. I do see what looks like a
battery in the memory section of the 1082 and it measures 0.08 VDC. Not a
good sign. I am really hoping someone has a Datron 1082 Service manual
with schematics that they can steer me to, preferably a soft copy so I can
better troubleshoot the unit. It might just need a new battery (plus a
re-cal) to get it going again but I hate going blind.
Randy
On Mon, Apr 21, 2014 at 8:15 AM, Tom Knox <actast@hotmail.com> wrote:
> As with many products of this age the tantalum caps are a common point of
> failure in the 1082.
> Although this current problem sounds like a memory battery.
>
> Thomas Knox
>
>
>
> > Date: Mon, 21 Apr 2014 11:03:56 -0400
> > From: cfharris@erols.com
> > To: volt-nuts@febo.com
> > Subject: Re: [volt-nuts] Datron 1082
> >
> > Just in case you haven't yet: Always start any investigation with
> > checking the power supply for voltage and ripple. Test the main
> > electrolytic capacitors for capacitance, and ESR.
> >
> > Nothing in the unit's self testing ability can be trusted if the
> > power supply is putting out large amounts of ripple, or is at the
> > wrong voltage.
> >
> > -Chuck Harris
> >
> > Randy Evans wrote:
> > > I have a Datron 1082 DMM that just recently stopped working. When I
> power
> > > up, it gives a "FAIL" message on the front panel. I found a flowchart
> in
> > > an abbreviated servicing manual that indicates it's from a failure of
> the
> > > "Check for uncorrupted calibration store". I Assume this likely means
> the
> > > cal tables are corrupted? Unfortunately, I don't have sufficient
> > > information (e.g., schematics) to do any further troubleshooting. Can
> > > anyone help with a complete servicing manual and/or further diagnostic
> help?
> > >
> > >
> > > I did download the Datron 1071DMM O&S manual from K04BB but I don't
> know
> > > how similar it is to the 1082. Can anyone comment on what the
> differences
> > > are?
> > >
> > >
> > >
> > > Thanks,
> > >
> > >
> > >
> > > Randy Evans
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
TK
Tom Knox
Tue, Apr 22, 2014 2:54 AM
Randy I would say from my experience with Datron there is about 90% chance the battery and recal will solve the problem. I would modify the battery with a coupe sets of leads that are long enough so you can install a new while the old battery is still in circuit.
Best Wishes;
Thomas Knox
Date: Mon, 21 Apr 2014 19:23:35 -0700
From: randyevans2688@gmail.com
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Datron 1082
I believe the voltages are correct but I have no schematics or
troubleshooting instructions to confirm. I do see what looks like a
battery in the memory section of the 1082 and it measures 0.08 VDC. Not a
good sign. I am really hoping someone has a Datron 1082 Service manual
with schematics that they can steer me to, preferably a soft copy so I can
better troubleshoot the unit. It might just need a new battery (plus a
re-cal) to get it going again but I hate going blind.
Randy
On Mon, Apr 21, 2014 at 8:15 AM, Tom Knox actast@hotmail.com wrote:
As with many products of this age the tantalum caps are a common point of
failure in the 1082.
Although this current problem sounds like a memory battery.
Thomas Knox
Date: Mon, 21 Apr 2014 11:03:56 -0400
From: cfharris@erols.com
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Datron 1082
Just in case you haven't yet: Always start any investigation with
checking the power supply for voltage and ripple. Test the main
electrolytic capacitors for capacitance, and ESR.
Nothing in the unit's self testing ability can be trusted if the
power supply is putting out large amounts of ripple, or is at the
wrong voltage.
-Chuck Harris
Randy Evans wrote:
I have a Datron 1082 DMM that just recently stopped working. When I
up, it gives a "FAIL" message on the front panel. I found a flowchart
an abbreviated servicing manual that indicates it's from a failure of
"Check for uncorrupted calibration store". I Assume this likely means
cal tables are corrupted? Unfortunately, I don't have sufficient
information (e.g., schematics) to do any further troubleshooting. Can
anyone help with a complete servicing manual and/or further diagnostic
I did download the Datron 1071DMM O&S manual from K04BB but I don't
how similar it is to the 1082. Can anyone comment on what the
and follow the instructions there.
Randy I would say from my experience with Datron there is about 90% chance the battery and recal will solve the problem. I would modify the battery with a coupe sets of leads that are long enough so you can install a new while the old battery is still in circuit.
Best Wishes;
Thomas Knox
> Date: Mon, 21 Apr 2014 19:23:35 -0700
> From: randyevans2688@gmail.com
> To: volt-nuts@febo.com
> Subject: Re: [volt-nuts] Datron 1082
>
> I believe the voltages are correct but I have no schematics or
> troubleshooting instructions to confirm. I do see what looks like a
> battery in the memory section of the 1082 and it measures 0.08 VDC. Not a
> good sign. I am really hoping someone has a Datron 1082 Service manual
> with schematics that they can steer me to, preferably a soft copy so I can
> better troubleshoot the unit. It might just need a new battery (plus a
> re-cal) to get it going again but I hate going blind.
>
> Randy
>
>
> On Mon, Apr 21, 2014 at 8:15 AM, Tom Knox <actast@hotmail.com> wrote:
>
> > As with many products of this age the tantalum caps are a common point of
> > failure in the 1082.
> > Although this current problem sounds like a memory battery.
> >
> > Thomas Knox
> >
> >
> >
> > > Date: Mon, 21 Apr 2014 11:03:56 -0400
> > > From: cfharris@erols.com
> > > To: volt-nuts@febo.com
> > > Subject: Re: [volt-nuts] Datron 1082
> > >
> > > Just in case you haven't yet: Always start any investigation with
> > > checking the power supply for voltage and ripple. Test the main
> > > electrolytic capacitors for capacitance, and ESR.
> > >
> > > Nothing in the unit's self testing ability can be trusted if the
> > > power supply is putting out large amounts of ripple, or is at the
> > > wrong voltage.
> > >
> > > -Chuck Harris
> > >
> > > Randy Evans wrote:
> > > > I have a Datron 1082 DMM that just recently stopped working. When I
> > power
> > > > up, it gives a "FAIL" message on the front panel. I found a flowchart
> > in
> > > > an abbreviated servicing manual that indicates it's from a failure of
> > the
> > > > "Check for uncorrupted calibration store". I Assume this likely means
> > the
> > > > cal tables are corrupted? Unfortunately, I don't have sufficient
> > > > information (e.g., schematics) to do any further troubleshooting. Can
> > > > anyone help with a complete servicing manual and/or further diagnostic
> > help?
> > > >
> > > >
> > > > I did download the Datron 1071DMM O&S manual from K04BB but I don't
> > know
> > > > how similar it is to the 1082. Can anyone comment on what the
> > differences
> > > > are?
> > > >
> > > >
> > > >
> > > > Thanks,
> > > >
> > > >
> > > >
> > > > Randy Evans
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> >
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
RE
Randy Evans
Tue, Apr 22, 2014 2:59 AM
Tom,
I would think with 0.08 VDC the battery is dead and all the memory is gone.
Do you think it would make any difference if I paralleled the leads with a
new one? I don't have a parts list to know what type the battery is so
it's going to be a challenge to find a direct replacement. Wish I had a
manual.
Randy
On Mon, Apr 21, 2014 at 7:54 PM, Tom Knox actast@hotmail.com wrote:
Randy I would say from my experience with Datron there is about 90% chance
the battery and recal will solve the problem. I would modify the battery
with a coupe sets of leads that are long enough so you can install a new
while the old battery is still in circuit.
Best Wishes;
Thomas Knox
Date: Mon, 21 Apr 2014 19:23:35 -0700
From: randyevans2688@gmail.com
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Datron 1082
I believe the voltages are correct but I have no schematics or
troubleshooting instructions to confirm. I do see what looks like a
battery in the memory section of the 1082 and it measures 0.08 VDC. Not a
good sign. I am really hoping someone has a Datron 1082 Service manual
with schematics that they can steer me to, preferably a soft copy so I
better troubleshoot the unit. It might just need a new battery (plus a
re-cal) to get it going again but I hate going blind.
Randy
On Mon, Apr 21, 2014 at 8:15 AM, Tom Knox actast@hotmail.com wrote:
As with many products of this age the tantalum caps are a common point
failure in the 1082.
Although this current problem sounds like a memory battery.
Thomas Knox
Date: Mon, 21 Apr 2014 11:03:56 -0400
From: cfharris@erols.com
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Datron 1082
Just in case you haven't yet: Always start any investigation with
checking the power supply for voltage and ripple. Test the main
electrolytic capacitors for capacitance, and ESR.
Nothing in the unit's self testing ability can be trusted if the
power supply is putting out large amounts of ripple, or is at the
wrong voltage.
-Chuck Harris
Randy Evans wrote:
I have a Datron 1082 DMM that just recently stopped working. When
up, it gives a "FAIL" message on the front panel. I found a
an abbreviated servicing manual that indicates it's from a failure
"Check for uncorrupted calibration store". I Assume this likely
cal tables are corrupted? Unfortunately, I don't have sufficient
information (e.g., schematics) to do any further troubleshooting.
anyone help with a complete servicing manual and/or further
I did download the Datron 1071DMM O&S manual from K04BB but I
how similar it is to the 1082. Can anyone comment on what the
and follow the instructions there.
and follow the instructions there.
Tom,
I would think with 0.08 VDC the battery is dead and all the memory is gone.
Do you think it would make any difference if I paralleled the leads with a
new one? I don't have a parts list to know what type the battery is so
it's going to be a challenge to find a direct replacement. Wish I had a
manual.
Randy
On Mon, Apr 21, 2014 at 7:54 PM, Tom Knox <actast@hotmail.com> wrote:
> Randy I would say from my experience with Datron there is about 90% chance
> the battery and recal will solve the problem. I would modify the battery
> with a coupe sets of leads that are long enough so you can install a new
> while the old battery is still in circuit.
> Best Wishes;
> Thomas Knox
>
>
>
> > Date: Mon, 21 Apr 2014 19:23:35 -0700
> > From: randyevans2688@gmail.com
> > To: volt-nuts@febo.com
> > Subject: Re: [volt-nuts] Datron 1082
> >
> > I believe the voltages are correct but I have no schematics or
> > troubleshooting instructions to confirm. I do see what looks like a
> > battery in the memory section of the 1082 and it measures 0.08 VDC. Not a
> > good sign. I am really hoping someone has a Datron 1082 Service manual
> > with schematics that they can steer me to, preferably a soft copy so I
> can
> > better troubleshoot the unit. It might just need a new battery (plus a
> > re-cal) to get it going again but I hate going blind.
> >
> > Randy
> >
> >
> > On Mon, Apr 21, 2014 at 8:15 AM, Tom Knox <actast@hotmail.com> wrote:
> >
> > > As with many products of this age the tantalum caps are a common point
> of
> > > failure in the 1082.
> > > Although this current problem sounds like a memory battery.
> > >
> > > Thomas Knox
> > >
> > >
> > >
> > > > Date: Mon, 21 Apr 2014 11:03:56 -0400
> > > > From: cfharris@erols.com
> > > > To: volt-nuts@febo.com
> > > > Subject: Re: [volt-nuts] Datron 1082
> > > >
> > > > Just in case you haven't yet: Always start any investigation with
> > > > checking the power supply for voltage and ripple. Test the main
> > > > electrolytic capacitors for capacitance, and ESR.
> > > >
> > > > Nothing in the unit's self testing ability can be trusted if the
> > > > power supply is putting out large amounts of ripple, or is at the
> > > > wrong voltage.
> > > >
> > > > -Chuck Harris
> > > >
> > > > Randy Evans wrote:
> > > > > I have a Datron 1082 DMM that just recently stopped working. When
> I
> > > power
> > > > > up, it gives a "FAIL" message on the front panel. I found a
> flowchart
> > > in
> > > > > an abbreviated servicing manual that indicates it's from a failure
> of
> > > the
> > > > > "Check for uncorrupted calibration store". I Assume this likely
> means
> > > the
> > > > > cal tables are corrupted? Unfortunately, I don't have sufficient
> > > > > information (e.g., schematics) to do any further troubleshooting.
> Can
> > > > > anyone help with a complete servicing manual and/or further
> diagnostic
> > > help?
> > > > >
> > > > >
> > > > > I did download the Datron 1071DMM O&S manual from K04BB but I
> don't
> > > know
> > > > > how similar it is to the 1082. Can anyone comment on what the
> > > differences
> > > > > are?
> > > > >
> > > > >
> > > > >
> > > > > Thanks,
> > > > >
> > > > >
> > > > >
> > > > > Randy Evans
> > > > _______________________________________________
> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
TK
Tom Knox
Tue, Apr 22, 2014 3:32 AM
Sorry I wasn't clear. I think the errors you are experiencing are because the memory battery is gone, so it won't matter now, this is for saving the cal data during future replacement to avoid these problems again.
Best Wishes;
Thomas Knox
Date: Mon, 21 Apr 2014 19:59:29 -0700
From: randyevans2688@gmail.com
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Datron 1082
Tom,
I would think with 0.08 VDC the battery is dead and all the memory is gone.
Do you think it would make any difference if I paralleled the leads with a
new one? I don't have a parts list to know what type the battery is so
it's going to be a challenge to find a direct replacement. Wish I had a
manual.
Randy
On Mon, Apr 21, 2014 at 7:54 PM, Tom Knox actast@hotmail.com wrote:
Randy I would say from my experience with Datron there is about 90% chance
the battery and recal will solve the problem. I would modify the battery
with a coupe sets of leads that are long enough so you can install a new
while the old battery is still in circuit.
Best Wishes;
Thomas Knox
Date: Mon, 21 Apr 2014 19:23:35 -0700
From: randyevans2688@gmail.com
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Datron 1082
I believe the voltages are correct but I have no schematics or
troubleshooting instructions to confirm. I do see what looks like a
battery in the memory section of the 1082 and it measures 0.08 VDC. Not a
good sign. I am really hoping someone has a Datron 1082 Service manual
with schematics that they can steer me to, preferably a soft copy so I
better troubleshoot the unit. It might just need a new battery (plus a
re-cal) to get it going again but I hate going blind.
Randy
On Mon, Apr 21, 2014 at 8:15 AM, Tom Knox actast@hotmail.com wrote:
As with many products of this age the tantalum caps are a common point
failure in the 1082.
Although this current problem sounds like a memory battery.
Thomas Knox
Date: Mon, 21 Apr 2014 11:03:56 -0400
From: cfharris@erols.com
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Datron 1082
Just in case you haven't yet: Always start any investigation with
checking the power supply for voltage and ripple. Test the main
electrolytic capacitors for capacitance, and ESR.
Nothing in the unit's self testing ability can be trusted if the
power supply is putting out large amounts of ripple, or is at the
wrong voltage.
-Chuck Harris
Randy Evans wrote:
I have a Datron 1082 DMM that just recently stopped working. When
up, it gives a "FAIL" message on the front panel. I found a
an abbreviated servicing manual that indicates it's from a failure
"Check for uncorrupted calibration store". I Assume this likely
cal tables are corrupted? Unfortunately, I don't have sufficient
information (e.g., schematics) to do any further troubleshooting.
anyone help with a complete servicing manual and/or further
I did download the Datron 1071DMM O&S manual from K04BB but I
how similar it is to the 1082. Can anyone comment on what the
and follow the instructions there.
and follow the instructions there.
Sorry I wasn't clear. I think the errors you are experiencing are because the memory battery is gone, so it won't matter now, this is for saving the cal data during future replacement to avoid these problems again.
Best Wishes;
Thomas Knox
> Date: Mon, 21 Apr 2014 19:59:29 -0700
> From: randyevans2688@gmail.com
> To: volt-nuts@febo.com
> Subject: Re: [volt-nuts] Datron 1082
>
> Tom,
>
> I would think with 0.08 VDC the battery is dead and all the memory is gone.
> Do you think it would make any difference if I paralleled the leads with a
> new one? I don't have a parts list to know what type the battery is so
> it's going to be a challenge to find a direct replacement. Wish I had a
> manual.
>
> Randy
>
>
> On Mon, Apr 21, 2014 at 7:54 PM, Tom Knox <actast@hotmail.com> wrote:
>
> > Randy I would say from my experience with Datron there is about 90% chance
> > the battery and recal will solve the problem. I would modify the battery
> > with a coupe sets of leads that are long enough so you can install a new
> > while the old battery is still in circuit.
> > Best Wishes;
> > Thomas Knox
> >
> >
> >
> > > Date: Mon, 21 Apr 2014 19:23:35 -0700
> > > From: randyevans2688@gmail.com
> > > To: volt-nuts@febo.com
> > > Subject: Re: [volt-nuts] Datron 1082
> > >
> > > I believe the voltages are correct but I have no schematics or
> > > troubleshooting instructions to confirm. I do see what looks like a
> > > battery in the memory section of the 1082 and it measures 0.08 VDC. Not a
> > > good sign. I am really hoping someone has a Datron 1082 Service manual
> > > with schematics that they can steer me to, preferably a soft copy so I
> > can
> > > better troubleshoot the unit. It might just need a new battery (plus a
> > > re-cal) to get it going again but I hate going blind.
> > >
> > > Randy
> > >
> > >
> > > On Mon, Apr 21, 2014 at 8:15 AM, Tom Knox <actast@hotmail.com> wrote:
> > >
> > > > As with many products of this age the tantalum caps are a common point
> > of
> > > > failure in the 1082.
> > > > Although this current problem sounds like a memory battery.
> > > >
> > > > Thomas Knox
> > > >
> > > >
> > > >
> > > > > Date: Mon, 21 Apr 2014 11:03:56 -0400
> > > > > From: cfharris@erols.com
> > > > > To: volt-nuts@febo.com
> > > > > Subject: Re: [volt-nuts] Datron 1082
> > > > >
> > > > > Just in case you haven't yet: Always start any investigation with
> > > > > checking the power supply for voltage and ripple. Test the main
> > > > > electrolytic capacitors for capacitance, and ESR.
> > > > >
> > > > > Nothing in the unit's self testing ability can be trusted if the
> > > > > power supply is putting out large amounts of ripple, or is at the
> > > > > wrong voltage.
> > > > >
> > > > > -Chuck Harris
> > > > >
> > > > > Randy Evans wrote:
> > > > > > I have a Datron 1082 DMM that just recently stopped working. When
> > I
> > > > power
> > > > > > up, it gives a "FAIL" message on the front panel. I found a
> > flowchart
> > > > in
> > > > > > an abbreviated servicing manual that indicates it's from a failure
> > of
> > > > the
> > > > > > "Check for uncorrupted calibration store". I Assume this likely
> > means
> > > > the
> > > > > > cal tables are corrupted? Unfortunately, I don't have sufficient
> > > > > > information (e.g., schematics) to do any further troubleshooting.
> > Can
> > > > > > anyone help with a complete servicing manual and/or further
> > diagnostic
> > > > help?
> > > > > >
> > > > > >
> > > > > > I did download the Datron 1071DMM O&S manual from K04BB but I
> > don't
> > > > know
> > > > > > how similar it is to the 1082. Can anyone comment on what the
> > > > differences
> > > > > > are?
> > > > > >
> > > > > >
> > > > > >
> > > > > > Thanks,
> > > > > >
> > > > > >
> > > > > >
> > > > > > Randy Evans
> > > > > _______________________________________________
> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > To unsubscribe, go to
> > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > and follow the instructions there.
> > > >
> > > > _______________________________________________
> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > To unsubscribe, go to
> > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> >
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
RE
Randy Evans
Tue, Apr 22, 2014 7:39 PM
The schematic of the Datron 1071 DMM seems to be very similar to the Datron
1082 but I'm sure there are differences. Unfortunately, I have not been
able to locate the 1082 Service Manual with the parts lists and schematics.
The 1071 manual does have a similar or identical battery but I can't find
the part number in the 1071 parts list. Does anyone know the part number
of the memory battery in the 1082 or 1071, or the characteristics; e.g.,
voltage and A-H rating?
Thanks,
Randy Evans AE6YG
On Mon, Apr 21, 2014 at 8:32 PM, Tom Knox actast@hotmail.com wrote:
Sorry I wasn't clear. I think the errors you are experiencing are because
the memory battery is gone, so it won't matter now, this is for saving the
cal data during future replacement to avoid these problems again.
Best Wishes;
Thomas Knox
Do you think it would make any difference if I paralleled the leads
new one? I don't have a parts list to know what type the battery is so
it's going to be a challenge to find a direct replacement. Wish I had a
manual.
Randy
On Mon, Apr 21, 2014 at 7:54 PM, Tom Knox actast@hotmail.com wrote:
Randy I would say from my experience with Datron there is about 90%
the battery and recal will solve the problem. I would modify the
with a coupe sets of leads that are long enough so you can install a
while the old battery is still in circuit.
Best Wishes;
Thomas Knox
Date: Mon, 21 Apr 2014 19:23:35 -0700
From: randyevans2688@gmail.com
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Datron 1082
I believe the voltages are correct but I have no schematics or
troubleshooting instructions to confirm. I do see what looks like a
battery in the memory section of the 1082 and it measures 0.08 VDC.
good sign. I am really hoping someone has a Datron 1082 Service
with schematics that they can steer me to, preferably a soft copy so
better troubleshoot the unit. It might just need a new battery
re-cal) to get it going again but I hate going blind.
Randy
On Mon, Apr 21, 2014 at 8:15 AM, Tom Knox actast@hotmail.com
As with many products of this age the tantalum caps are a common
failure in the 1082.
Although this current problem sounds like a memory battery.
Thomas Knox
Date: Mon, 21 Apr 2014 11:03:56 -0400
From: cfharris@erols.com
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Datron 1082
Just in case you haven't yet: Always start any investigation with
checking the power supply for voltage and ripple. Test the main
electrolytic capacitors for capacitance, and ESR.
Nothing in the unit's self testing ability can be trusted if the
power supply is putting out large amounts of ripple, or is at the
wrong voltage.
-Chuck Harris
Randy Evans wrote:
I have a Datron 1082 DMM that just recently stopped working.
up, it gives a "FAIL" message on the front panel. I found a
an abbreviated servicing manual that indicates it's from a
"Check for uncorrupted calibration store". I Assume this
cal tables are corrupted? Unfortunately, I don't have
information (e.g., schematics) to do any further
anyone help with a complete servicing manual and/or further
I did download the Datron 1071DMM O&S manual from K04BB but I
how similar it is to the 1082. Can anyone comment on what the
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
The schematic of the Datron 1071 DMM seems to be very similar to the Datron
1082 but I'm sure there are differences. Unfortunately, I have not been
able to locate the 1082 Service Manual with the parts lists and schematics.
The 1071 manual does have a similar or identical battery but I can't find
the part number in the 1071 parts list. Does anyone know the part number
of the memory battery in the 1082 or 1071, or the characteristics; e.g.,
voltage and A-H rating?
Thanks,
Randy Evans AE6YG
On Mon, Apr 21, 2014 at 8:32 PM, Tom Knox <actast@hotmail.com> wrote:
> Sorry I wasn't clear. I think the errors you are experiencing are because
> the memory battery is gone, so it won't matter now, this is for saving the
> cal data during future replacement to avoid these problems again.
> Best Wishes;
> Thomas Knox
>
>
>
> > Date: Mon, 21 Apr 2014 19:59:29 -0700
> > From: randyevans2688@gmail.com
> > To: volt-nuts@febo.com
> > Subject: Re: [volt-nuts] Datron 1082
> >
> > Tom,
> >
> > I would think with 0.08 VDC the battery is dead and all the memory is
> gone.
> > Do you think it would make any difference if I paralleled the leads
> with a
> > new one? I don't have a parts list to know what type the battery is so
> > it's going to be a challenge to find a direct replacement. Wish I had a
> > manual.
> >
> > Randy
> >
> >
> > On Mon, Apr 21, 2014 at 7:54 PM, Tom Knox <actast@hotmail.com> wrote:
> >
> > > Randy I would say from my experience with Datron there is about 90%
> chance
> > > the battery and recal will solve the problem. I would modify the
> battery
> > > with a coupe sets of leads that are long enough so you can install a
> new
> > > while the old battery is still in circuit.
> > > Best Wishes;
> > > Thomas Knox
> > >
> > >
> > >
> > > > Date: Mon, 21 Apr 2014 19:23:35 -0700
> > > > From: randyevans2688@gmail.com
> > > > To: volt-nuts@febo.com
> > > > Subject: Re: [volt-nuts] Datron 1082
> > > >
> > > > I believe the voltages are correct but I have no schematics or
> > > > troubleshooting instructions to confirm. I do see what looks like a
> > > > battery in the memory section of the 1082 and it measures 0.08 VDC.
> Not a
> > > > good sign. I am really hoping someone has a Datron 1082 Service
> manual
> > > > with schematics that they can steer me to, preferably a soft copy so
> I
> > > can
> > > > better troubleshoot the unit. It might just need a new battery
> (plus a
> > > > re-cal) to get it going again but I hate going blind.
> > > >
> > > > Randy
> > > >
> > > >
> > > > On Mon, Apr 21, 2014 at 8:15 AM, Tom Knox <actast@hotmail.com>
> wrote:
> > > >
> > > > > As with many products of this age the tantalum caps are a common
> point
> > > of
> > > > > failure in the 1082.
> > > > > Although this current problem sounds like a memory battery.
> > > > >
> > > > > Thomas Knox
> > > > >
> > > > >
> > > > >
> > > > > > Date: Mon, 21 Apr 2014 11:03:56 -0400
> > > > > > From: cfharris@erols.com
> > > > > > To: volt-nuts@febo.com
> > > > > > Subject: Re: [volt-nuts] Datron 1082
> > > > > >
> > > > > > Just in case you haven't yet: Always start any investigation with
> > > > > > checking the power supply for voltage and ripple. Test the main
> > > > > > electrolytic capacitors for capacitance, and ESR.
> > > > > >
> > > > > > Nothing in the unit's self testing ability can be trusted if the
> > > > > > power supply is putting out large amounts of ripple, or is at the
> > > > > > wrong voltage.
> > > > > >
> > > > > > -Chuck Harris
> > > > > >
> > > > > > Randy Evans wrote:
> > > > > > > I have a Datron 1082 DMM that just recently stopped working.
> When
> > > I
> > > > > power
> > > > > > > up, it gives a "FAIL" message on the front panel. I found a
> > > flowchart
> > > > > in
> > > > > > > an abbreviated servicing manual that indicates it's from a
> failure
> > > of
> > > > > the
> > > > > > > "Check for uncorrupted calibration store". I Assume this
> likely
> > > means
> > > > > the
> > > > > > > cal tables are corrupted? Unfortunately, I don't have
> sufficient
> > > > > > > information (e.g., schematics) to do any further
> troubleshooting.
> > > Can
> > > > > > > anyone help with a complete servicing manual and/or further
> > > diagnostic
> > > > > help?
> > > > > > >
> > > > > > >
> > > > > > > I did download the Datron 1071DMM O&S manual from K04BB but I
> > > don't
> > > > > know
> > > > > > > how similar it is to the 1082. Can anyone comment on what the
> > > > > differences
> > > > > > > are?
> > > > > > >
> > > > > > >
> > > > > > >
> > > > > > > Thanks,
> > > > > > >
> > > > > > >
> > > > > > >
> > > > > > > Randy Evans
> > > > > > _______________________________________________
> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > > To unsubscribe, go to
> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > > and follow the instructions there.
> > > > >
> > > > > _______________________________________________
> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > To unsubscribe, go to
> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > and follow the instructions there.
> > > > >
> > > > _______________________________________________
> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
TM
Tom Miller
Tue, Apr 22, 2014 7:54 PM
Can you post a picture of the battery somewhere and what is printed on it?
Tom
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Tuesday, April 22, 2014 3:39 PM
Subject: Re: [volt-nuts] Datron 1082
The schematic of the Datron 1071 DMM seems to be very similar to the
Datron
1082 but I'm sure there are differences. Unfortunately, I have not been
able to locate the 1082 Service Manual with the parts lists and
schematics.
The 1071 manual does have a similar or identical battery but I can't find
the part number in the 1071 parts list. Does anyone know the part number
of the memory battery in the 1082 or 1071, or the characteristics; e.g.,
voltage and A-H rating?
Thanks,
Randy Evans AE6YG
On Mon, Apr 21, 2014 at 8:32 PM, Tom Knox actast@hotmail.com wrote:
Sorry I wasn't clear. I think the errors you are experiencing are because
the memory battery is gone, so it won't matter now, this is for saving
the
cal data during future replacement to avoid these problems again.
Best Wishes;
Thomas Knox
Do you think it would make any difference if I paralleled the leads
new one? I don't have a parts list to know what type the battery is so
it's going to be a challenge to find a direct replacement. Wish I had
a
manual.
Randy
On Mon, Apr 21, 2014 at 7:54 PM, Tom Knox actast@hotmail.com wrote:
Randy I would say from my experience with Datron there is about 90%
the battery and recal will solve the problem. I would modify the
with a coupe sets of leads that are long enough so you can install a
while the old battery is still in circuit.
Best Wishes;
Thomas Knox
Date: Mon, 21 Apr 2014 19:23:35 -0700
From: randyevans2688@gmail.com
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Datron 1082
I believe the voltages are correct but I have no schematics or
troubleshooting instructions to confirm. I do see what looks like
a
battery in the memory section of the 1082 and it measures 0.08 VDC.
good sign. I am really hoping someone has a Datron 1082 Service
with schematics that they can steer me to, preferably a soft copy
so
better troubleshoot the unit. It might just need a new battery
re-cal) to get it going again but I hate going blind.
Randy
On Mon, Apr 21, 2014 at 8:15 AM, Tom Knox actast@hotmail.com
As with many products of this age the tantalum caps are a common
failure in the 1082.
Although this current problem sounds like a memory battery.
Thomas Knox
Date: Mon, 21 Apr 2014 11:03:56 -0400
From: cfharris@erols.com
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Datron 1082
Just in case you haven't yet: Always start any investigation
with
checking the power supply for voltage and ripple. Test the
main
electrolytic capacitors for capacitance, and ESR.
Nothing in the unit's self testing ability can be trusted if
the
power supply is putting out large amounts of ripple, or is at
the
wrong voltage.
-Chuck Harris
Randy Evans wrote:
I have a Datron 1082 DMM that just recently stopped working.
up, it gives a "FAIL" message on the front panel. I found a
an abbreviated servicing manual that indicates it's from a
"Check for uncorrupted calibration store". I Assume this
cal tables are corrupted? Unfortunately, I don't have
information (e.g., schematics) to do any further
anyone help with a complete servicing manual and/or further
I did download the Datron 1071DMM O&S manual from K04BB but
I
how similar it is to the 1082. Can anyone comment on what
the
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
Can you post a picture of the battery somewhere and what is printed on it?
Tom
----- Original Message -----
From: "Randy Evans" <randyevans2688@gmail.com>
To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
Sent: Tuesday, April 22, 2014 3:39 PM
Subject: Re: [volt-nuts] Datron 1082
> The schematic of the Datron 1071 DMM seems to be very similar to the
> Datron
> 1082 but I'm sure there are differences. Unfortunately, I have not been
> able to locate the 1082 Service Manual with the parts lists and
> schematics.
> The 1071 manual does have a similar or identical battery but I can't find
> the part number in the 1071 parts list. Does anyone know the part number
> of the memory battery in the 1082 or 1071, or the characteristics; e.g.,
> voltage and A-H rating?
>
> Thanks,
>
> Randy Evans AE6YG
>
>
> On Mon, Apr 21, 2014 at 8:32 PM, Tom Knox <actast@hotmail.com> wrote:
>
>> Sorry I wasn't clear. I think the errors you are experiencing are because
>> the memory battery is gone, so it won't matter now, this is for saving
>> the
>> cal data during future replacement to avoid these problems again.
>> Best Wishes;
>> Thomas Knox
>>
>>
>>
>> > Date: Mon, 21 Apr 2014 19:59:29 -0700
>> > From: randyevans2688@gmail.com
>> > To: volt-nuts@febo.com
>> > Subject: Re: [volt-nuts] Datron 1082
>> >
>> > Tom,
>> >
>> > I would think with 0.08 VDC the battery is dead and all the memory is
>> gone.
>> > Do you think it would make any difference if I paralleled the leads
>> with a
>> > new one? I don't have a parts list to know what type the battery is so
>> > it's going to be a challenge to find a direct replacement. Wish I had
>> > a
>> > manual.
>> >
>> > Randy
>> >
>> >
>> > On Mon, Apr 21, 2014 at 7:54 PM, Tom Knox <actast@hotmail.com> wrote:
>> >
>> > > Randy I would say from my experience with Datron there is about 90%
>> chance
>> > > the battery and recal will solve the problem. I would modify the
>> battery
>> > > with a coupe sets of leads that are long enough so you can install a
>> new
>> > > while the old battery is still in circuit.
>> > > Best Wishes;
>> > > Thomas Knox
>> > >
>> > >
>> > >
>> > > > Date: Mon, 21 Apr 2014 19:23:35 -0700
>> > > > From: randyevans2688@gmail.com
>> > > > To: volt-nuts@febo.com
>> > > > Subject: Re: [volt-nuts] Datron 1082
>> > > >
>> > > > I believe the voltages are correct but I have no schematics or
>> > > > troubleshooting instructions to confirm. I do see what looks like
>> > > > a
>> > > > battery in the memory section of the 1082 and it measures 0.08 VDC.
>> Not a
>> > > > good sign. I am really hoping someone has a Datron 1082 Service
>> manual
>> > > > with schematics that they can steer me to, preferably a soft copy
>> > > > so
>> I
>> > > can
>> > > > better troubleshoot the unit. It might just need a new battery
>> (plus a
>> > > > re-cal) to get it going again but I hate going blind.
>> > > >
>> > > > Randy
>> > > >
>> > > >
>> > > > On Mon, Apr 21, 2014 at 8:15 AM, Tom Knox <actast@hotmail.com>
>> wrote:
>> > > >
>> > > > > As with many products of this age the tantalum caps are a common
>> point
>> > > of
>> > > > > failure in the 1082.
>> > > > > Although this current problem sounds like a memory battery.
>> > > > >
>> > > > > Thomas Knox
>> > > > >
>> > > > >
>> > > > >
>> > > > > > Date: Mon, 21 Apr 2014 11:03:56 -0400
>> > > > > > From: cfharris@erols.com
>> > > > > > To: volt-nuts@febo.com
>> > > > > > Subject: Re: [volt-nuts] Datron 1082
>> > > > > >
>> > > > > > Just in case you haven't yet: Always start any investigation
>> > > > > > with
>> > > > > > checking the power supply for voltage and ripple. Test the
>> > > > > > main
>> > > > > > electrolytic capacitors for capacitance, and ESR.
>> > > > > >
>> > > > > > Nothing in the unit's self testing ability can be trusted if
>> > > > > > the
>> > > > > > power supply is putting out large amounts of ripple, or is at
>> > > > > > the
>> > > > > > wrong voltage.
>> > > > > >
>> > > > > > -Chuck Harris
>> > > > > >
>> > > > > > Randy Evans wrote:
>> > > > > > > I have a Datron 1082 DMM that just recently stopped working.
>> When
>> > > I
>> > > > > power
>> > > > > > > up, it gives a "FAIL" message on the front panel. I found a
>> > > flowchart
>> > > > > in
>> > > > > > > an abbreviated servicing manual that indicates it's from a
>> failure
>> > > of
>> > > > > the
>> > > > > > > "Check for uncorrupted calibration store". I Assume this
>> likely
>> > > means
>> > > > > the
>> > > > > > > cal tables are corrupted? Unfortunately, I don't have
>> sufficient
>> > > > > > > information (e.g., schematics) to do any further
>> troubleshooting.
>> > > Can
>> > > > > > > anyone help with a complete servicing manual and/or further
>> > > diagnostic
>> > > > > help?
>> > > > > > >
>> > > > > > >
>> > > > > > > I did download the Datron 1071DMM O&S manual from K04BB but
>> > > > > > > I
>> > > don't
>> > > > > know
>> > > > > > > how similar it is to the 1082. Can anyone comment on what
>> > > > > > > the
>> > > > > differences
>> > > > > > > are?
>> > > > > > >
>> > > > > > >
>> > > > > > >
>> > > > > > > Thanks,
>> > > > > > >
>> > > > > > >
>> > > > > > >
>> > > > > > > Randy Evans
>> > > > > > _______________________________________________
>> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > > > > To unsubscribe, go to
>> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > > > > > and follow the instructions there.
>> > > > >
>> > > > > _______________________________________________
>> > > > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > > > To unsubscribe, go to
>> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > > > > and follow the instructions there.
>> > > > >
>> > > > _______________________________________________
>> > > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > > To unsubscribe, go to
>> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > > > and follow the instructions there.
>> > >
>> > > _______________________________________________
>> > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > To unsubscribe, go to
>> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > > and follow the instructions there.
>> > >
>> > _______________________________________________
>> > volt-nuts mailing list -- volt-nuts@febo.com
>> > To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > and follow the instructions there.
>>
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
TM
T. Micallef
Wed, Apr 23, 2014 2:07 AM
Randy Evans <randyevans2688@...> writes:
The schematic of the Datron 1071 DMM seems to be very similar to the Datron
1082 but I'm sure there are differences. Unfortunately, I have not been
able to locate the 1082 Service Manual with the parts lists and schematics.
The 1071 manual does have a similar or identical battery but I can't find
the part number in the 1071 parts list. Does anyone know the part number
of the memory battery in the 1082 or 1071, or the characteristics; e.g.,
voltage and A-H rating?
Thanks,
Randy Evans AE6YG
On Mon, Apr 21, 2014 at 8:32 PM, Tom Knox <actast@...> wrote:
Randy Evans <randyevans2688@...> writes:
>
> The schematic of the Datron 1071 DMM seems to be very similar to the Datron
> 1082 but I'm sure there are differences. Unfortunately, I have not been
> able to locate the 1082 Service Manual with the parts lists and schematics.
> The 1071 manual does have a similar or identical battery but I can't find
> the part number in the 1071 parts list. Does anyone know the part number
> of the memory battery in the 1082 or 1071, or the characteristics; e.g.,
> voltage and A-H rating?
>
> Thanks,
>
> Randy Evans AE6YG
>
> On Mon, Apr 21, 2014 at 8:32 PM, Tom Knox <actast@...> wrote:
>
I popped the top off of my mothballed 1071 and it looks like a 1/2AA battery
based on physical measurements. It looks like the one in the following link
http://www.zbattery.com/Tadiran-TL-5151-with-tabs-1-2-AA-Lithium-Battery
Todd