On Samstag, 5. Februar 2022 10:48:57 CET Julien Goodwin wrote:

I /think/ it's using more power than it did before (right around 20W at
total cold, dipping to 10W after a while), although silly me didn't make
a note of it, but even after 30 minutes it's not locking.

The power starting high and dipping eventually means there's some regulation
for the temperature. That's a good sign.

Given that the resistor I replaced is in a key point for the lamp
excitation oscillator I won't be surprised if it's totally off and needs
a recalibration, but that's a task for another day. When I stick a scope
probe in the lamp cavity as the manual suggests I do see a a signal, so
there's certainly hope, although I don't see the expected glow.

First goal is to get the lamp to light up, once that's working, "calibration"
of the lamp exciter is only fine tuning. To see if the exciter oscillator is
starting at all you could use a makeshift RF probe with an oscilloscope and a
coil of wire connected to the probe and the get that close to the lamp. You
should see something between 70 and 90 MHz if the oscillator is working.

If it isn't, hopefully Q2 isn't damaged. Voltage across R14 should tell. There
should be a voltage drop across it, maybe 1V or so? Lower if the oscillator is
not working, but certainly not 0. If the voltage is 0V, Q2 has either failed
open or the connection to the exciter coil is broken. If the voltage is much
higher, Q2 has probably failed shorted and then R14 should also get very, very
hot.

I did confirm the heater is working, and was getting above 90c when I
poked a thermocouple at it just after power-off.

That's quite low. Nominally, the lamp temperature should be 113°C.

The good thing about the FRK is the very detailed service manual that tells
you exact troubleshooting steps for each of the component. I guess you've
already found it?

BR,
Matthias

On Samstag, 5. Februar 2022 10:48:57 CET Julien Goodwin wrote: > I /think/ it's using more power than it did before (right around 20W at > total cold, dipping to 10W after a while), although silly me didn't make > a note of it, but even after 30 minutes it's not locking. The power starting high and dipping eventually means there's some regulation for the temperature. That's a good sign. > Given that the resistor I replaced is in a key point for the lamp > excitation oscillator I won't be surprised if it's totally off and needs > a recalibration, but that's a task for another day. When I stick a scope > probe in the lamp cavity as the manual suggests I do see a a signal, so > there's certainly hope, although I don't see the expected glow. First goal is to get the lamp to light up, once that's working, "calibration" of the lamp exciter is only fine tuning. To see if the exciter oscillator is starting at all you could use a makeshift RF probe with an oscilloscope and a coil of wire connected to the probe and the get that close to the lamp. You should see something between 70 and 90 MHz if the oscillator is working. If it isn't, hopefully Q2 isn't damaged. Voltage across R14 should tell. There should be a voltage drop across it, maybe 1V or so? Lower if the oscillator is not working, but certainly not 0. If the voltage is 0V, Q2 has either failed open or the connection to the exciter coil is broken. If the voltage is much higher, Q2 has probably failed shorted and then R14 should also get very, very hot. > I did confirm the heater is working, and was getting above 90c when I > poked a thermocouple at it just after power-off. That's quite low. Nominally, the lamp temperature should be 113°C. The good thing about the FRK is the very detailed service manual that tells you exact troubleshooting steps for each of the component. I guess you've already found it? BR, Matthias
JG
Julien Goodwin
Fri, Feb 11, 2022 5:17 AM

On 6/2/22 1:14 am, Matthias Welwarsky wrote:

On Samstag, 5. Februar 2022 10:48:57 CET Julien Goodwin wrote:

I /think/ it's using more power than it did before (right around 20W at
total cold, dipping to 10W after a while), although silly me didn't make
a note of it, but even after 30 minutes it's not locking.

The power starting high and dipping eventually means there's some regulation
for the temperature. That's a good sign.

Given that the resistor I replaced is in a key point for the lamp
excitation oscillator I won't be surprised if it's totally off and needs
a recalibration, but that's a task for another day. When I stick a scope
probe in the lamp cavity as the manual suggests I do see a a signal, so
there's certainly hope, although I don't see the expected glow.

First goal is to get the lamp to light up, once that's working, "calibration"
of the lamp exciter is only fine tuning. To see if the exciter oscillator is
starting at all you could use a makeshift RF probe with an oscilloscope and a
coil of wire connected to the probe and the get that close to the lamp. You
should see something between 70 and 90 MHz if the oscillator is working.

I did stick a scope probe in as I mentioned, and there's certainly
oscillation now happening, although I didn't get a clear indication of
the frequency, tuning that is now second on the todo list once I sort
out the heater, and acquire a slightly safer way of probing that area
than waving a bare scope probe around (some field probes have been
ordered which should be ideal for this).

If it isn't, hopefully Q2 isn't damaged. Voltage across R14 should tell. There
should be a voltage drop across it, maybe 1V or so? Lower if the oscillator is
not working, but certainly not 0. If the voltage is 0V, Q2 has either failed
open or the connection to the exciter coil is broken. If the voltage is much
higher, Q2 has probably failed shorted and then R14 should also get very, very
hot.

I did confirm the heater is working, and was getting above 90c when I
poked a thermocouple at it just after power-off.

That's quite low. Nominally, the lamp temperature should be 113°C.

The good thing about the FRK is the very detailed service manual that tells
you exact troubleshooting steps for each of the component. I guess you've
already found it?

Yep, although I find myself switching between the FRK (mid-70s) and
FRK-L (late-80s) manuals as neither quite matches the device I have. I
guess that dates my device to likely early/mid-80s.

Until today I hadn't actually measured the temperature properly, and not
at all in operation, so I wasn't too worried that I was only seeing 90c.
Today I (kapton) taped a thermocouple to the outside lamp housing and
let it warm up for a while, and it ended up stabilizing still around
90c. So my earlier measurements were actually right on, sadly.

Some schematic & board review later it seems clear that someone's
reworked this at some point, the trimmer that should control the
temperature has been replaced with a fixed resistor, which also explains
the weird wires attached to the heater transistor.

Next time I do an order from my parts supplier (which, naturally I just
did the other day, oh well) I shall try and find something close enough,
as, naturally, a direct replacement is out of stock. Options that should
be close enough are there, so hopefully it'll be fine.

Sadly that means a full disassembly again, hopefully I don't lose too
many more screws this time, but at least I'll know to sort the work
surface out much better to protect against it.

On 6/2/22 1:14 am, Matthias Welwarsky wrote: > On Samstag, 5. Februar 2022 10:48:57 CET Julien Goodwin wrote: >> I /think/ it's using more power than it did before (right around 20W at >> total cold, dipping to 10W after a while), although silly me didn't make >> a note of it, but even after 30 minutes it's not locking. > > The power starting high and dipping eventually means there's some regulation > for the temperature. That's a good sign. > >> Given that the resistor I replaced is in a key point for the lamp >> excitation oscillator I won't be surprised if it's totally off and needs >> a recalibration, but that's a task for another day. When I stick a scope >> probe in the lamp cavity as the manual suggests I do see a a signal, so >> there's certainly hope, although I don't see the expected glow. > > First goal is to get the lamp to light up, once that's working, "calibration" > of the lamp exciter is only fine tuning. To see if the exciter oscillator is > starting at all you could use a makeshift RF probe with an oscilloscope and a > coil of wire connected to the probe and the get that close to the lamp. You > should see something between 70 and 90 MHz if the oscillator is working. I did stick a scope probe in as I mentioned, and there's certainly oscillation now happening, although I didn't get a clear indication of the frequency, tuning that is now second on the todo list once I sort out the heater, and acquire a slightly safer way of probing that area than waving a bare scope probe around (some field probes have been ordered which should be ideal for this). > If it isn't, hopefully Q2 isn't damaged. Voltage across R14 should tell. There > should be a voltage drop across it, maybe 1V or so? Lower if the oscillator is > not working, but certainly not 0. If the voltage is 0V, Q2 has either failed > open or the connection to the exciter coil is broken. If the voltage is much > higher, Q2 has probably failed shorted and then R14 should also get very, very > hot. > >> I did confirm the heater is working, and was getting above 90c when I >> poked a thermocouple at it just after power-off. > > That's quite low. Nominally, the lamp temperature should be 113°C. > > The good thing about the FRK is the very detailed service manual that tells > you exact troubleshooting steps for each of the component. I guess you've > already found it? Yep, although I find myself switching between the FRK (mid-70s) and FRK-L (late-80s) manuals as neither quite matches the device I have. I guess that dates my device to likely early/mid-80s. Until today I hadn't actually measured the temperature properly, and not at all in operation, so I wasn't too worried that I was only seeing 90c. Today I (kapton) taped a thermocouple to the outside lamp housing and let it warm up for a while, and it ended up stabilizing still around 90c. So my earlier measurements were actually right on, sadly. Some schematic & board review later it seems clear that someone's reworked this at some point, the trimmer that should control the temperature has been replaced with a fixed resistor, which also explains the weird wires attached to the heater transistor. Next time I do an order from my parts supplier (which, naturally I just did the other day, oh well) I shall try and find something close enough, as, naturally, a direct replacement is out of stock. Options that should be close enough are there, so hopefully it'll be fine. Sadly that means a full disassembly again, hopefully I don't lose too many more screws this time, but at least I'll know to sort the work surface out much better to protect against it.
PS
paul swed
Fri, Feb 11, 2022 7:34 PM

With respect to the lamp oscillator the frequency is not that particular.
Not sure you need to adjust it if the lamp ignites. Think they are in the
137 MHz region.
Good luck
Paul
WB8TSL

On Fri, Feb 11, 2022 at 1:31 AM Julien Goodwin time-nuts@studio442.com.au
wrote:

On 6/2/22 1:14 am, Matthias Welwarsky wrote:

On Samstag, 5. Februar 2022 10:48:57 CET Julien Goodwin wrote:

I /think/ it's using more power than it did before (right around 20W at
total cold, dipping to 10W after a while), although silly me didn't make
a note of it, but even after 30 minutes it's not locking.

The power starting high and dipping eventually means there's some

regulation

for the temperature. That's a good sign.

Given that the resistor I replaced is in a key point for the lamp
excitation oscillator I won't be surprised if it's totally off and needs
a recalibration, but that's a task for another day. When I stick a scope
probe in the lamp cavity as the manual suggests I do see a a signal, so
there's certainly hope, although I don't see the expected glow.

First goal is to get the lamp to light up, once that's working,

"calibration"

of the lamp exciter is only fine tuning. To see if the exciter

oscillator is

starting at all you could use a makeshift RF probe with an oscilloscope

and a

coil of wire connected to the probe and the get that close to the lamp.

You

should see something between 70 and 90 MHz if the oscillator is working.

I did stick a scope probe in as I mentioned, and there's certainly
oscillation now happening, although I didn't get a clear indication of
the frequency, tuning that is now second on the todo list once I sort
out the heater, and acquire a slightly safer way of probing that area
than waving a bare scope probe around (some field probes have been
ordered which should be ideal for this).

If it isn't, hopefully Q2 isn't damaged. Voltage across R14 should tell.

There

should be a voltage drop across it, maybe 1V or so? Lower if the

oscillator is

not working, but certainly not 0. If the voltage is 0V, Q2 has either

failed

open or the connection to the exciter coil is broken. If the voltage is

much

higher, Q2 has probably failed shorted and then R14 should also get

very, very

hot.

I did confirm the heater is working, and was getting above 90c when I
poked a thermocouple at it just after power-off.

That's quite low. Nominally, the lamp temperature should be 113°C.

The good thing about the FRK is the very detailed service manual that

tells

you exact troubleshooting steps for each of the component. I guess you've
already found it?

Yep, although I find myself switching between the FRK (mid-70s) and
FRK-L (late-80s) manuals as neither quite matches the device I have. I
guess that dates my device to likely early/mid-80s.

Until today I hadn't actually measured the temperature properly, and not
at all in operation, so I wasn't too worried that I was only seeing 90c.
Today I (kapton) taped a thermocouple to the outside lamp housing and
let it warm up for a while, and it ended up stabilizing still around
90c. So my earlier measurements were actually right on, sadly.

Some schematic & board review later it seems clear that someone's
reworked this at some point, the trimmer that should control the
temperature has been replaced with a fixed resistor, which also explains
the weird wires attached to the heater transistor.

Next time I do an order from my parts supplier (which, naturally I just
did the other day, oh well) I shall try and find something close enough,
as, naturally, a direct replacement is out of stock. Options that should
be close enough are there, so hopefully it'll be fine.

Sadly that means a full disassembly again, hopefully I don't lose too
many more screws this time, but at least I'll know to sort the work
surface out much better to protect against it.


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

With respect to the lamp oscillator the frequency is not that particular. Not sure you need to adjust it if the lamp ignites. Think they are in the 137 MHz region. Good luck Paul WB8TSL On Fri, Feb 11, 2022 at 1:31 AM Julien Goodwin <time-nuts@studio442.com.au> wrote: > > > On 6/2/22 1:14 am, Matthias Welwarsky wrote: > > On Samstag, 5. Februar 2022 10:48:57 CET Julien Goodwin wrote: > >> I /think/ it's using more power than it did before (right around 20W at > >> total cold, dipping to 10W after a while), although silly me didn't make > >> a note of it, but even after 30 minutes it's not locking. > > > > The power starting high and dipping eventually means there's some > regulation > > for the temperature. That's a good sign. > > > >> Given that the resistor I replaced is in a key point for the lamp > >> excitation oscillator I won't be surprised if it's totally off and needs > >> a recalibration, but that's a task for another day. When I stick a scope > >> probe in the lamp cavity as the manual suggests I do see a a signal, so > >> there's certainly hope, although I don't see the expected glow. > > > > First goal is to get the lamp to light up, once that's working, > "calibration" > > of the lamp exciter is only fine tuning. To see if the exciter > oscillator is > > starting at all you could use a makeshift RF probe with an oscilloscope > and a > > coil of wire connected to the probe and the get that close to the lamp. > You > > should see something between 70 and 90 MHz if the oscillator is working. > > I did stick a scope probe in as I mentioned, and there's certainly > oscillation now happening, although I didn't get a clear indication of > the frequency, tuning that is now second on the todo list once I sort > out the heater, and acquire a slightly safer way of probing that area > than waving a bare scope probe around (some field probes have been > ordered which should be ideal for this). > > > If it isn't, hopefully Q2 isn't damaged. Voltage across R14 should tell. > There > > should be a voltage drop across it, maybe 1V or so? Lower if the > oscillator is > > not working, but certainly not 0. If the voltage is 0V, Q2 has either > failed > > open or the connection to the exciter coil is broken. If the voltage is > much > > higher, Q2 has probably failed shorted and then R14 should also get > very, very > > hot. > > > >> I did confirm the heater is working, and was getting above 90c when I > >> poked a thermocouple at it just after power-off. > > > > That's quite low. Nominally, the lamp temperature should be 113°C. > > > > The good thing about the FRK is the very detailed service manual that > tells > > you exact troubleshooting steps for each of the component. I guess you've > > already found it? > > Yep, although I find myself switching between the FRK (mid-70s) and > FRK-L (late-80s) manuals as neither quite matches the device I have. I > guess that dates my device to likely early/mid-80s. > > Until today I hadn't actually measured the temperature properly, and not > at all in operation, so I wasn't too worried that I was only seeing 90c. > Today I (kapton) taped a thermocouple to the outside lamp housing and > let it warm up for a while, and it ended up stabilizing still around > 90c. So my earlier measurements were actually right on, sadly. > > Some schematic & board review later it seems clear that someone's > reworked this at some point, the trimmer that should control the > temperature has been replaced with a fixed resistor, which also explains > the weird wires attached to the heater transistor. > > Next time I do an order from my parts supplier (which, naturally I just > did the other day, oh well) I shall try and find something close enough, > as, naturally, a direct replacement is out of stock. Options that should > be close enough are there, so hopefully it'll be fine. > > Sadly that means a full disassembly again, hopefully I don't lose too > many more screws this time, but at least I'll know to sort the work > surface out much better to protect against it. > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send > an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there.
DF
Darren Freeman
Tue, Feb 15, 2022 7:10 AM

On Sat, 2022-02-05 at 20:48 +1100, Julien Goodwin wrote:

I /think/ it's using more power than it did before (right around 20W
at
total cold, dipping to 10W after a while), although silly me didn't
make
a note of it, but even after 30 minutes it's not locking.

I don't have any experience with this brand/model, but I have
resurrected some "dead" rubidium standards. I'm mostly going from
experience with gas lasers, but I found the same approach worked on the
rubidium units that I have here.

It's well worth being patient with the lamp, if it's been off for many
years, it may have a gas pressure/mix that's far from optimal. If you
let it sit with the heater and exciter running, it may just strike
after a while.

(You should measure the current going into the exciter, it should jump
up suddenly when the lamp strikes. Measure the unlit condition with the
lamp cold and the exciter on. On mine, the current roughly doubles when
it is lit.)

If it does strike, keep it running for a few days, don't power it off.
It can get even harder to light, before it gets easier again. You might
consider using a UPS for this procedure, as Murphy could strike just
after the lamp does!

On one of my units, I had to manually run the lamp heater to a higher
than normal temperature, before it would light. (With a thermocouple,
it's best to use some heatsink grease, otherwise the readings can be a
long way off.)

The needed temperature gradually came down, until the unit could be
reassembled and used normally. On another, I just waited for several
hours at the usual temperature, and that was enough. Although again, it
took a couple of days of operation before it would light without
hesitation.

On yet another unit, the lamp would light normally, but the photodiode
signal was too weak to achieve a lock. (The firmware said nope.) So I
just left it running overnight, and then after a power cycle, it locked
immediately!

So I wanted to point out that if the lamp has everything it needs, and
still won't light, you might just need to leave it powered for a while.
And you have not much to lose, if you need to run it at a higher
temperature for a while.

Have fun,
Darren

On Sat, 2022-02-05 at 20:48 +1100, Julien Goodwin wrote: > I /think/ it's using more power than it did before (right around 20W > at > total cold, dipping to 10W after a while), although silly me didn't > make > a note of it, but even after 30 minutes it's not locking. I don't have any experience with this brand/model, but I have resurrected some "dead" rubidium standards. I'm mostly going from experience with gas lasers, but I found the same approach worked on the rubidium units that I have here. It's well worth being patient with the lamp, if it's been off for many years, it may have a gas pressure/mix that's far from optimal. If you let it sit with the heater and exciter running, it may just strike after a while. (You should measure the current going into the exciter, it should jump up suddenly when the lamp strikes. Measure the unlit condition with the lamp cold and the exciter on. On mine, the current roughly doubles when it is lit.) If it does strike, keep it running for a few days, don't power it off. It can get even harder to light, before it gets easier again. You might consider using a UPS for this procedure, as Murphy could strike just after the lamp does! On one of my units, I had to manually run the lamp heater to a higher than normal temperature, before it would light. (With a thermocouple, it's best to use some heatsink grease, otherwise the readings can be a long way off.) The needed temperature gradually came down, until the unit could be reassembled and used normally. On another, I just waited for several hours at the usual temperature, and that was enough. Although again, it took a couple of days of operation before it would light without hesitation. On yet another unit, the lamp would light normally, but the photodiode signal was too weak to achieve a lock. (The firmware said nope.) So I just left it running overnight, and then after a power cycle, it locked immediately! So I wanted to point out that if the lamp has everything it needs, and still won't light, you might just need to leave it powered for a while. And you have not much to lose, if you need to run it at a higher temperature for a while. Have fun, Darren
MW
Matthias Welwarsky
Tue, Feb 15, 2022 5:17 PM

On Dienstag, 15. Februar 2022 08:10:12 CET Darren Freeman wrote:

On one of my units, I had to manually run the lamp heater to a higher
than normal temperature, before it would light. (With a thermocouple,
it's best to use some heatsink grease, otherwise the readings can be a
long way off.)

The needed temperature gradually came down, until the unit could be
reassembled and used normally. On another, I just waited for several
hours at the usual temperature, and that was enough. Although again, it
took a couple of days of operation before it would light without
hesitation.

On the same note, if you have a hot air rework station, you can try gently
heating the lamp assembly from a distance, it may help getting the lamp to
ignite, too.

On Dienstag, 15. Februar 2022 08:10:12 CET Darren Freeman wrote: > On one of my units, I had to manually run the lamp heater to a higher > than normal temperature, before it would light. (With a thermocouple, > it's best to use some heatsink grease, otherwise the readings can be a > long way off.) > > The needed temperature gradually came down, until the unit could be > reassembled and used normally. On another, I just waited for several > hours at the usual temperature, and that was enough. Although again, it > took a couple of days of operation before it would light without > hesitation. On the same note, if you have a hot air rework station, you can try gently heating the lamp assembly from a distance, it may help getting the lamp to ignite, too.
JG
Julien Goodwin
Fri, Feb 25, 2022 8:10 AM

This afternoon I swapped the fixed resistor that should have been a
trimpot back to a trimpot, and quickly dialed it up to where I was
getting ~115c on the lamp surface as expected.

As I was in there anyway I also swapped the 20 ohm resistor back to the
Vishay wirewound part it should have been from the hack I'd made up
using cheap carbon resistors from the local shop.

No obvious light, but I also still need to see if the excitation
oscillator is even vaguely in the ballpark (or possibly working, I may
potentially have messed with things while I was in there today), for
which I'm still waiting on some safer probe options.

On 15/2/22 6:10 pm, Darren Freeman wrote:

On Sat, 2022-02-05 at 20:48 +1100, Julien Goodwin wrote:

I /think/ it's using more power than it did before (right around 20W
at
total cold, dipping to 10W after a while), although silly me didn't
make
a note of it, but even after 30 minutes it's not locking.

I don't have any experience with this brand/model, but I have
resurrected some "dead" rubidium standards. I'm mostly going from
experience with gas lasers, but I found the same approach worked on the
rubidium units that I have here.

It's well worth being patient with the lamp, if it's been off for many
years, it may have a gas pressure/mix that's far from optimal. If you
let it sit with the heater and exciter running, it may just strike
after a while.

I need to replace the main filter cap in the FRT power supply before
I'll be comfortable leaving it on for an extended duration, but once I
do (and the replacement is now here, I just need to hook it up) I'll be
fine leaving it on for a week or so to see.

If it doesn't light after that, I may try just swapping in the lamp from
my house standard FRK-LN to see if it's that simple, and this has a bad
lamp (at which point I think I'd give up for now).

(You should measure the current going into the exciter, it should jump
up suddenly when the lamp strikes. Measure the unlit condition with the
lamp cold and the exciter on. On mine, the current roughly doubles when
it is lit.)

If it does strike, keep it running for a few days, don't power it off.
It can get even harder to light, before it gets easier again. You might
consider using a UPS for this procedure, as Murphy could strike just
after the lamp does!

Fortunately Sydney power is good enough that my UPS' have caused more
outages than they've helped with.

On one of my units, I had to manually run the lamp heater to a higher
than normal temperature, before it would light. (With a thermocouple,
it's best to use some heatsink grease, otherwise the readings can be a
long way off.)

The needed temperature gradually came down, until the unit could be
reassembled and used normally. On another, I just waited for several
hours at the usual temperature, and that was enough. Although again, it
took a couple of days of operation before it would light without
hesitation.

On yet another unit, the lamp would light normally, but the photodiode
signal was too weak to achieve a lock. (The firmware said nope.) So I
just left it running overnight, and then after a power cycle, it locked
immediately!

So I wanted to point out that if the lamp has everything it needs, and
still won't light, you might just need to leave it powered for a while.
And you have not much to lose, if you need to run it at a higher
temperature for a while.

Have fun,
Darren


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

This afternoon I swapped the fixed resistor that should have been a trimpot back to a trimpot, and quickly dialed it up to where I was getting ~115c on the lamp surface as expected. As I was in there anyway I also swapped the 20 ohm resistor back to the Vishay wirewound part it should have been from the hack I'd made up using cheap carbon resistors from the local shop. No obvious light, but I also still need to see if the excitation oscillator is even vaguely in the ballpark (or possibly working, I may potentially have messed with things while I was in there today), for which I'm still waiting on some safer probe options. On 15/2/22 6:10 pm, Darren Freeman wrote: > On Sat, 2022-02-05 at 20:48 +1100, Julien Goodwin wrote: >> I /think/ it's using more power than it did before (right around 20W >> at >> total cold, dipping to 10W after a while), although silly me didn't >> make >> a note of it, but even after 30 minutes it's not locking. > > I don't have any experience with this brand/model, but I have > resurrected some "dead" rubidium standards. I'm mostly going from > experience with gas lasers, but I found the same approach worked on the > rubidium units that I have here. > > It's well worth being patient with the lamp, if it's been off for many > years, it may have a gas pressure/mix that's far from optimal. If you > let it sit with the heater and exciter running, it may just strike > after a while. I need to replace the main filter cap in the FRT power supply before I'll be comfortable leaving it on for an extended duration, but once I do (and the replacement is now here, I just need to hook it up) I'll be fine leaving it on for a week or so to see. If it doesn't light after that, I may try just swapping in the lamp from my house standard FRK-LN to see if it's that simple, and this has a bad lamp (at which point I think I'd give up for now). > (You should measure the current going into the exciter, it should jump > up suddenly when the lamp strikes. Measure the unlit condition with the > lamp cold and the exciter on. On mine, the current roughly doubles when > it is lit.) > > If it does strike, keep it running for a few days, don't power it off. > It can get even harder to light, before it gets easier again. You might > consider using a UPS for this procedure, as Murphy could strike just > after the lamp does! Fortunately Sydney power is good enough that my UPS' have caused more outages than they've helped with. > On one of my units, I had to manually run the lamp heater to a higher > than normal temperature, before it would light. (With a thermocouple, > it's best to use some heatsink grease, otherwise the readings can be a > long way off.) > > The needed temperature gradually came down, until the unit could be > reassembled and used normally. On another, I just waited for several > hours at the usual temperature, and that was enough. Although again, it > took a couple of days of operation before it would light without > hesitation. > > On yet another unit, the lamp would light normally, but the photodiode > signal was too weak to achieve a lock. (The firmware said nope.) So I > just left it running overnight, and then after a power cycle, it locked > immediately! > > So I wanted to point out that if the lamp has everything it needs, and > still won't light, you might just need to leave it powered for a while. > And you have not much to lose, if you need to run it at a higher > temperature for a while. > > Have fun, > Darren > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there.
BK
Bob kb8tq
Fri, Feb 25, 2022 1:32 PM

Hi

The oscillator is a “works or it doesn’t” sort of thing. Frequency and output
level usually are close enough to do what is needed.

Bob

On Feb 25, 2022, at 3:10 AM, Julien Goodwin time-nuts@studio442.com.au wrote:

This afternoon I swapped the fixed resistor that should have been a trimpot back to a trimpot, and quickly dialed it up to where I was getting ~115c on the lamp surface as expected.

As I was in there anyway I also swapped the 20 ohm resistor back to the Vishay wirewound part it should have been from the hack I'd made up using cheap carbon resistors from the local shop.

No obvious light, but I also still need to see if the excitation oscillator is even vaguely in the ballpark (or possibly working, I may potentially have messed with things while I was in there today), for which I'm still waiting on some safer probe options.

On 15/2/22 6:10 pm, Darren Freeman wrote:

On Sat, 2022-02-05 at 20:48 +1100, Julien Goodwin wrote:

I /think/ it's using more power than it did before (right around 20W
at
total cold, dipping to 10W after a while), although silly me didn't
make
a note of it, but even after 30 minutes it's not locking.

I don't have any experience with this brand/model, but I have
resurrected some "dead" rubidium standards. I'm mostly going from
experience with gas lasers, but I found the same approach worked on the
rubidium units that I have here.
It's well worth being patient with the lamp, if it's been off for many
years, it may have a gas pressure/mix that's far from optimal. If you
let it sit with the heater and exciter running, it may just strike
after a while.

I need to replace the main filter cap in the FRT power supply before I'll be comfortable leaving it on for an extended duration, but once I do (and the replacement is now here, I just need to hook it up) I'll be fine leaving it on for a week or so to see.

If it doesn't light after that, I may try just swapping in the lamp from my house standard FRK-LN to see if it's that simple, and this has a bad lamp (at which point I think I'd give up for now).

(You should measure the current going into the exciter, it should jump
up suddenly when the lamp strikes. Measure the unlit condition with the
lamp cold and the exciter on. On mine, the current roughly doubles when
it is lit.)
If it does strike, keep it running for a few days, don't power it off.
It can get even harder to light, before it gets easier again. You might
consider using a UPS for this procedure, as Murphy could strike just
after the lamp does!

Fortunately Sydney power is good enough that my UPS' have caused more outages than they've helped with.

On one of my units, I had to manually run the lamp heater to a higher
than normal temperature, before it would light. (With a thermocouple,
it's best to use some heatsink grease, otherwise the readings can be a
long way off.)
The needed temperature gradually came down, until the unit could be
reassembled and used normally. On another, I just waited for several
hours at the usual temperature, and that was enough. Although again, it
took a couple of days of operation before it would light without
hesitation.
On yet another unit, the lamp would light normally, but the photodiode
signal was too weak to achieve a lock. (The firmware said nope.) So I
just left it running overnight, and then after a power cycle, it locked
immediately!
So I wanted to point out that if the lamp has everything it needs, and
still won't light, you might just need to leave it powered for a while.
And you have not much to lose, if you need to run it at a higher
temperature for a while.
Have fun,
Darren


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

Hi The oscillator is a “works or it doesn’t” sort of thing. Frequency and output level usually are close enough to do what is needed. Bob > On Feb 25, 2022, at 3:10 AM, Julien Goodwin <time-nuts@studio442.com.au> wrote: > > This afternoon I swapped the fixed resistor that should have been a trimpot back to a trimpot, and quickly dialed it up to where I was getting ~115c on the lamp surface as expected. > > As I was in there anyway I also swapped the 20 ohm resistor back to the Vishay wirewound part it should have been from the hack I'd made up using cheap carbon resistors from the local shop. > > No obvious light, but I also still need to see if the excitation oscillator is even vaguely in the ballpark (or possibly working, I may potentially have messed with things while I was in there today), for which I'm still waiting on some safer probe options. > > On 15/2/22 6:10 pm, Darren Freeman wrote: >> On Sat, 2022-02-05 at 20:48 +1100, Julien Goodwin wrote: >>> I /think/ it's using more power than it did before (right around 20W >>> at >>> total cold, dipping to 10W after a while), although silly me didn't >>> make >>> a note of it, but even after 30 minutes it's not locking. >> I don't have any experience with this brand/model, but I have >> resurrected some "dead" rubidium standards. I'm mostly going from >> experience with gas lasers, but I found the same approach worked on the >> rubidium units that I have here. >> It's well worth being patient with the lamp, if it's been off for many >> years, it may have a gas pressure/mix that's far from optimal. If you >> let it sit with the heater and exciter running, it may just strike >> after a while. > > I need to replace the main filter cap in the FRT power supply before I'll be comfortable leaving it on for an extended duration, but once I do (and the replacement is now here, I just need to hook it up) I'll be fine leaving it on for a week or so to see. > > If it doesn't light after that, I may try just swapping in the lamp from my house standard FRK-LN to see if it's that simple, and this has a bad lamp (at which point I think I'd give up for now). > >> (You should measure the current going into the exciter, it should jump >> up suddenly when the lamp strikes. Measure the unlit condition with the >> lamp cold and the exciter on. On mine, the current roughly doubles when >> it is lit.) >> If it does strike, keep it running for a few days, don't power it off. >> It can get even harder to light, before it gets easier again. You might >> consider using a UPS for this procedure, as Murphy could strike just >> after the lamp does! > > Fortunately Sydney power is good enough that my UPS' have caused more outages than they've helped with. > >> On one of my units, I had to manually run the lamp heater to a higher >> than normal temperature, before it would light. (With a thermocouple, >> it's best to use some heatsink grease, otherwise the readings can be a >> long way off.) >> The needed temperature gradually came down, until the unit could be >> reassembled and used normally. On another, I just waited for several >> hours at the usual temperature, and that was enough. Although again, it >> took a couple of days of operation before it would light without >> hesitation. >> On yet another unit, the lamp would light normally, but the photodiode >> signal was too weak to achieve a lock. (The firmware said nope.) So I >> just left it running overnight, and then after a power cycle, it locked >> immediately! >> So I wanted to point out that if the lamp has everything it needs, and >> still won't light, you might just need to leave it powered for a while. >> And you have not much to lose, if you need to run it at a higher >> temperature for a while. >> Have fun, >> Darren >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com >> To unsubscribe, go to and follow the instructions there. > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there.