time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

Re: Types of noise (was: Phase Station 53100A Questions)

JG
Joseph Gwinn
Tue, Feb 15, 2022 10:07 PM

On Tue, 15 Feb 2022 03:30:31 -0500, time-nuts-request@lists.febo.com
wrote:
time-nuts Digest, Vol 214, Issue 17

Date: Mon, 14 Feb 2022 22:10:10 +0100
From: Attila Kinali attila@kinali.ch
Subject: [time-nuts] Re: Types of noise (was: Phase Station 53100A
Questions)
To: Discussion of precise time and frequency measurement
time-nuts@lists.febo.com
Message-ID: 20220214221010.7d39f07da972b13bc6900de7@kinali.ch
Content-Type: text/plain; charset=UTF-8

On Sun, 13 Feb 2022 17:31:11 -0500
Joseph Gwinn joegwinn@comcast.net wrote:

A better word than multiplicative is parametric, the varying
parameters being path loss and path group delay.  This is as seen at
the phase noise test set.

That is one way how noise can enter a system. Not the only one though.
And once it is in the system, non-linear elements can increase it.
Look at the derivations in [1] (and to a lesser extend [2]), which
shows how the non-linearity of an element leads to up and down conversion
of noise and thus turns additive noise into multiplicative noise.
Unfortunately, whatever measurement system you use, will have non-linear
elements in it. Starting from the input amplifier and going to the
sampling system. So this kind of noise amplification is unavoidable.
(Yes, I know that I am blowing my own horn, but these are the only
publications that I am aware of that explain this phenomena at all.
Yes, I find this odd too.)

Yes, and I read these with great interest when they came out.

Amplitude and phase noise are looking at noise from two different
perspective. One is how large the variation of the peak of a sine
wave is, the other is how much the zero crossing varies in time.
Note that all natural noise sources will be both amplitude and
phase noise.

Hmm.  One case I'm interested in is where the path attenuation varies
according to a random telegraph waveform, due to for instance a loose
connector or cracked center conductor rattling under heavy
vibration.

That's an interesting noise model. And one that is oddly specific.
Did you see that in some application? If yes, could you explain a bit more?

When integrating large systems, people have a lot of trouble with
loose and/or broken coax connectors, especially under vibration, or
when a cable is moved for some unrelated reason.  I'm looking for a
simple way to detect from the outside if this is happening, without
disassembling everything (and introducing added problems).  Nor is
the problem necessarily obvious to the eye, even if it is all
disassembled.

As discussed herein, I suspect that looking at AM (Amplitude
Modulation) PN (Phase Noise) and comparing it with PM (Phase
Modulation) PN will flag such problems.

I've also been reading the 200-page 53100A User Manual (dated 3 March
2021), and it turns out that the measurement hardware does collect
amplitude difference data, one assumes in a manner paralleling the
collection of phase difference data (page 38 therein).  The
strip-chart format is available for phase difference data, and an
amplitude strip chart could allow one to see rattling directly.  This
graph type is not currently implemented, but at least the data may be
available in the TIM file.

In this, the electrical length does not change.

I would suggest to be careful with this assumption. 1ps is equivalent
to 140µm to 200µm of distance in a cable/connector. And with phase
noise, we are looking at effects that have the equivalent time-deviation
of a few fs or a few 100nm of distance. Even if the contacts do not
move much,
they still move. And that will have all kinds of effects (including change
of capacitive coupling and thus impedance and thus reflection).
I would suggest you do some back of the envelope calculation to see
whether that movement would have some measurable effect on your system
or not.

Yes, I think that the cracked center conductor, a common fault, is
most simply modeled as a tiny series capacitor (perhaps coaxial) that
may be shorted (when no break), with 50ohm lines on either side, so
there may be very little length change.

But my instinct is that the path loss change effect will generally be
far larger, so we may still have a large excess of AM over PM when
something is rattling.  When I'm in the lab, I'll test the issue
directly, equipment permitting.

While
the source of the carrier whose PN is being measured will have some
mixture of AM and PM characteristic of that source, the residual
(added) PN will be characteristic of the transit damage encountered
between source and PN test set.  So wouldn't this randomly varying
attenuation yield mostly residual AM PN and little residual PM PN?

What is the PN here? Phase noise? If so, then there is no AM PN.

Well, Rubiola uses AM PN and PM PN, and TimeLab says Phase Noise and
AM Noise, and others use other names.  Plain PN was always a
misnomer, as there was always an AM component, even though PM
generally dominated.  Except in laser systems, where AM PN is also
known as RIN (Relative Intensity Noise), and PM PN was basically
unmeasurable.  I like the AM PN and PM PN nomenclature because it is
precise and symmetric, from DC to daylight.

Amplitude modulation modulates the amplitude. The zero crossing
does not change. Thus there is no phase noise. Phase noise comes
from phase modulation (or equivalently time modulation).

Yes.

Sure, there are elements that turn AM into PM and vice versa, but
that is a different topic and I don't want to complicate the
discussion more than necessary. Though with the on/off modulation
you have, that might be something you need to look into.

I was talking about the effect of AGC on AM PN in the
rattling-connection case.  My point was that AGC didn't help there.

I am not quite sure what kind of system you are trying to measure
and what you hope to see there. So I'm a bit careful in suggesting
things until know better what your setup is.

See example at beginning, from radar integration.  My focus is
transit damage, independent of source PN properties and issues.

Joe Gwinn

[1] "A Physical Sine-to-Square Converter Noise Model"
http://people.mpi-inf.mpg.de/~adogan/pubs/IFCS2018_comparator_noise.pdf

[2] "A Fresh Look at the Design of Low Jitter Hard Limiters"
http://people.mpi-inf.mpg.de/~adogan/pubs/IFCS2019_collins_isf.pdf

The driving force behind research is the question: "Why?"
There are things we don't understand and things we always
wonder about. And that's why we do research.
-- Kobayashi Makoto

Proceeded by something to the effect of "That's odd", not Eureka.

On Tue, 15 Feb 2022 03:30:31 -0500, time-nuts-request@lists.febo.com wrote: time-nuts Digest, Vol 214, Issue 17 > Date: Mon, 14 Feb 2022 22:10:10 +0100 > From: Attila Kinali <attila@kinali.ch> > Subject: [time-nuts] Re: Types of noise (was: Phase Station 53100A > Questions) > To: Discussion of precise time and frequency measurement > <time-nuts@lists.febo.com> > Message-ID: <20220214221010.7d39f07da972b13bc6900de7@kinali.ch> > Content-Type: text/plain; charset=UTF-8 > > On Sun, 13 Feb 2022 17:31:11 -0500 > Joseph Gwinn <joegwinn@comcast.net> wrote: > >> A better word than multiplicative is parametric, the varying >> parameters being path loss and path group delay. This is as seen at >> the phase noise test set. > > That is one way how noise can enter a system. Not the only one though. > And once it is in the system, non-linear elements can increase it. > Look at the derivations in [1] (and to a lesser extend [2]), which > shows how the non-linearity of an element leads to up and down conversion > of noise and thus turns additive noise into multiplicative noise. > Unfortunately, whatever measurement system you use, will have non-linear > elements in it. Starting from the input amplifier and going to the > sampling system. So this kind of noise amplification is unavoidable. > (Yes, I know that I am blowing my own horn, but these are the only > publications that I am aware of that explain this phenomena at all. > Yes, I find this odd too.) Yes, and I read these with great interest when they came out. >>> Amplitude and phase noise are looking at noise from two different >>> perspective. One is how large the variation of the peak of a sine >>> wave is, the other is how much the zero crossing varies in time. >>> Note that all natural noise sources will be both amplitude and >>> phase noise. >> >> Hmm. One case I'm interested in is where the path attenuation varies >> according to a random telegraph waveform, due to for instance a loose >> connector or cracked center conductor rattling under heavy >> vibration. > > That's an interesting noise model. And one that is oddly specific. > Did you see that in some application? If yes, could you explain a bit more? When integrating large systems, people have a lot of trouble with loose and/or broken coax connectors, especially under vibration, or when a cable is moved for some unrelated reason. I'm looking for a simple way to detect from the outside if this is happening, without disassembling everything (and introducing added problems). Nor is the problem necessarily obvious to the eye, even if it is all disassembled. As discussed herein, I suspect that looking at AM (Amplitude Modulation) PN (Phase Noise) and comparing it with PM (Phase Modulation) PN will flag such problems. I've also been reading the 200-page 53100A User Manual (dated 3 March 2021), and it turns out that the measurement hardware does collect amplitude difference data, one assumes in a manner paralleling the collection of phase difference data (page 38 therein). The strip-chart format is available for phase difference data, and an amplitude strip chart could allow one to see rattling directly. This graph type is not currently implemented, but at least the data may be available in the TIM file. >> In this, the electrical length does not change. > > I would suggest to be careful with this assumption. 1ps is equivalent > to 140µm to 200µm of distance in a cable/connector. And with phase > noise, we are looking at effects that have the equivalent time-deviation > of a few fs or a few 100nm of distance. Even if the contacts do not > move much, > they still move. And that will have all kinds of effects (including change > of capacitive coupling and thus impedance and thus reflection). > I would suggest you do some back of the envelope calculation to see > whether that movement would have some measurable effect on your system > or not. Yes, I think that the cracked center conductor, a common fault, is most simply modeled as a tiny series capacitor (perhaps coaxial) that may be shorted (when no break), with 50ohm lines on either side, so there may be very little length change. But my instinct is that the path loss change effect will generally be far larger, so we may still have a large excess of AM over PM when something is rattling. When I'm in the lab, I'll test the issue directly, equipment permitting. >> While >> the source of the carrier whose PN is being measured will have some >> mixture of AM and PM characteristic of that source, the residual >> (added) PN will be characteristic of the transit damage encountered >> between source and PN test set. So wouldn't this randomly varying >> attenuation yield mostly residual AM PN and little residual PM PN? > > What is the PN here? Phase noise? If so, then there is no AM PN. Well, Rubiola uses AM PN and PM PN, and TimeLab says Phase Noise and AM Noise, and others use other names. Plain PN was always a misnomer, as there was always an AM component, even though PM generally dominated. Except in laser systems, where AM PN is also known as RIN (Relative Intensity Noise), and PM PN was basically unmeasurable. I like the AM PN and PM PN nomenclature because it is precise and symmetric, from DC to daylight. > Amplitude modulation modulates the amplitude. The zero crossing > does not change. Thus there is no phase noise. Phase noise comes > from phase modulation (or equivalently time modulation). Yes. > Sure, there are elements that turn AM into PM and vice versa, but > that is a different topic and I don't want to complicate the > discussion more than necessary. Though with the on/off modulation > you have, that might be something you need to look into. I was talking about the effect of AGC on AM PN in the rattling-connection case. My point was that AGC didn't help there. > I am not quite sure what kind of system you are trying to measure > and what you hope to see there. So I'm a bit careful in suggesting > things until know better what your setup is. See example at beginning, from radar integration. My focus is transit damage, independent of source PN properties and issues. Joe Gwinn > > [1] "A Physical Sine-to-Square Converter Noise Model" > http://people.mpi-inf.mpg.de/~adogan/pubs/IFCS2018_comparator_noise.pdf > > [2] "A Fresh Look at the Design of Low Jitter Hard Limiters" > http://people.mpi-inf.mpg.de/~adogan/pubs/IFCS2019_collins_isf.pdf > -- > The driving force behind research is the question: "Why?" > There are things we don't understand and things we always > wonder about. And that's why we do research. > -- Kobayashi Makoto Proceeded by something to the effect of "That's odd", not Eureka.
AK
Attila Kinali
Fri, Feb 18, 2022 9:42 PM

On Tue, 15 Feb 2022 17:07:54 -0500
Joseph Gwinn joegwinn@comcast.net wrote:

That's an interesting noise model. And one that is oddly specific.
Did you see that in some application? If yes, could you explain a bit more?

When integrating large systems, people have a lot of trouble with
loose and/or broken coax connectors, especially under vibration, or
when a cable is moved for some unrelated reason.  I'm looking for a
simple way to detect from the outside if this is happening, without
disassembling everything (and introducing added problems).  Nor is
the problem necessarily obvious to the eye, even if it is all
disassembled.

Hmm... And what makes you think that phase noise would be the best
way to do so? Wouldn't looking at the impedance match at the sending
side, i.e. looking at the reflected wave be the better approach?
That should be a much better method and could potentially tell you
also where the problem is. (this is being used in large scale
fiber networks to detect breaks)

While
the source of the carrier whose PN is being measured will have some
mixture of AM and PM characteristic of that source, the residual
(added) PN will be characteristic of the transit damage encountered
between source and PN test set.  So wouldn't this randomly varying
attenuation yield mostly residual AM PN and little residual PM PN?

What is the PN here? Phase noise? If so, then there is no AM PN.

Well, Rubiola uses AM PN and PM PN,

Does he? I am not aware of that. Quite to the contrary, in his
talks and presentations he does make sure that people do not
confuse AM and PM.

If you look at [1], specifically at slide 3 and 12 you will see
that he splits total noise on a signal into AM and PM noise.
I.e. total noise is the sum of AM and PM noise.

and TimeLab says Phase Noise and
AM Noise, and others use other names.  Plain PN was always a
misnomer, as there was always an AM component, even though PM
generally dominated.

I think what you mean here by PN is total noise, not phase noise.

Except in laser systems, where AM PN is also
known as RIN (Relative Intensity Noise), and PM PN was basically
unmeasurable.  I like the AM PN and PM PN nomenclature because it is
precise and symmetric, from DC to daylight.

Ok.. there is some confusion here. RIN is just an other word for
amplitude noise. The reason why lasers have (had) such a high amplitude
noise is because, being oscillators with no active amplitude control,
that are based on a stochastic process with a high positive gain but only
a slow process to restore energy into the system, the fluctuations of
photon density in the cavity are basically uncontrolled and jump up
and down quite a lot. But because the photon density never sinks to zero,
the phase is preserved. But mind you, high AM noise (aka RIN) translates
to increased PM noise as well, due to AM-PM translation in the (non-linear)
gain medium (laser are very weird devices that are plagued by all kinds
of problems if you are looking too closely).

As for PM noise being unmeasureable, that's not true. It was and is
measurable. It was just that we did not have the tools to do so
commonly available until a few decades ago. To do phase noise
measurament of lasers, you need a fast photo diode to beat two
lasers against each other, then you analyse the output signal with
a phase noise analyser. Coincidentally, with the photo diode you
get both AM and PM noise.

Today, narrow line-width lasers are all phase noise (or rather
frequency noise) controlled, additional to being amplitude controlled.
The three major techniques for this are saturated absorption spectroscopy,
dichroic atomic vapor lock and Pound-Drever-Hall lock.
Sub-Hz line-widths are quite common and the best get down to ~50mHz.

		Attila Kinali

[1] "Basics of Phase Noise" by Enrico Rubiola, IFCS/PTTI 2005 tutorial
http://rubiola.org/pdf-slides/2005T-ifcs-basics.pdf

The driving force behind research is the question: "Why?"
There are things we don't understand and things we always
wonder about. And that's why we do research.
-- Kobayashi Makoto

On Tue, 15 Feb 2022 17:07:54 -0500 Joseph Gwinn <joegwinn@comcast.net> wrote: > > That's an interesting noise model. And one that is oddly specific. > > Did you see that in some application? If yes, could you explain a bit more? > > When integrating large systems, people have a lot of trouble with > loose and/or broken coax connectors, especially under vibration, or > when a cable is moved for some unrelated reason. I'm looking for a > simple way to detect from the outside if this is happening, without > disassembling everything (and introducing added problems). Nor is > the problem necessarily obvious to the eye, even if it is all > disassembled. Hmm... And what makes you think that phase noise would be the best way to do so? Wouldn't looking at the impedance match at the sending side, i.e. looking at the reflected wave be the better approach? That should be a much better method and could potentially tell you also where the problem is. (this is being used in large scale fiber networks to detect breaks) > >> While > >> the source of the carrier whose PN is being measured will have some > >> mixture of AM and PM characteristic of that source, the residual > >> (added) PN will be characteristic of the transit damage encountered > >> between source and PN test set. So wouldn't this randomly varying > >> attenuation yield mostly residual AM PN and little residual PM PN? > > > > What is the PN here? Phase noise? If so, then there is no AM PN. > > Well, Rubiola uses AM PN and PM PN, Does he? I am not aware of that. Quite to the contrary, in his talks and presentations he does make sure that people do not confuse AM and PM. If you look at [1], specifically at slide 3 and 12 you will see that he splits total noise on a signal into AM and PM noise. I.e. total noise is the sum of AM and PM noise. > and TimeLab says Phase Noise and > AM Noise, and others use other names. Plain PN was always a > misnomer, as there was always an AM component, even though PM > generally dominated. I think what you mean here by PN is total noise, not phase noise. > Except in laser systems, where AM PN is also > known as RIN (Relative Intensity Noise), and PM PN was basically > unmeasurable. I like the AM PN and PM PN nomenclature because it is > precise and symmetric, from DC to daylight. Ok.. there is some confusion here. RIN is just an other word for amplitude noise. The reason why lasers have (had) such a high amplitude noise is because, being oscillators with no active amplitude control, that are based on a stochastic process with a high positive gain but only a slow process to restore energy into the system, the fluctuations of photon density in the cavity are basically uncontrolled and jump up and down quite a lot. But because the photon density never sinks to zero, the phase is preserved. But mind you, high AM noise (aka RIN) translates to increased PM noise as well, due to AM-PM translation in the (non-linear) gain medium (laser are very weird devices that are plagued by all kinds of problems if you are looking too closely). As for PM noise being unmeasureable, that's not true. It was and is measurable. It was just that we did not have the tools to do so commonly available until a few decades ago. To do phase noise measurament of lasers, you need a fast photo diode to beat two lasers against each other, then you analyse the output signal with a phase noise analyser. Coincidentally, with the photo diode you get both AM and PM noise. Today, narrow line-width lasers are all phase noise (or rather frequency noise) controlled, additional to being amplitude controlled. The three major techniques for this are saturated absorption spectroscopy, dichroic atomic vapor lock and Pound-Drever-Hall lock. Sub-Hz line-widths are quite common and the best get down to ~50mHz. Attila Kinali [1] "Basics of Phase Noise" by Enrico Rubiola, IFCS/PTTI 2005 tutorial http://rubiola.org/pdf-slides/2005T-ifcs-basics.pdf -- The driving force behind research is the question: "Why?" There are things we don't understand and things we always wonder about. And that's why we do research. -- Kobayashi Makoto