Dear Group,
Good morning! I just got my first HP-53132A frequency counter and Prologix GPIB-USB converter. Now I can do some measurements and I am excited about it. Everything looks promising. Condition of the instrument on the outside is like new, controls responsive and no fading in lettering at the buttons. Connectors are not worn or damaged, original connectors. The unit has been tested and works OK, probably not with adequate calibration but that is not an issue now, I can always calibrate it later.
What worries me is the display. Even though being an Agilent built unit, it has probably more hours on it than a seldom used HP built unit. The flourescent display is fading, not all the digits show consistent illumination levels. The left-most digits show some flickering at some of their segments. It is readable in the end, but I am terribly disappointed about it.
Does anybody know if there is a cure for this? Or should I start saving for a better unit.
I am not afraid of the screwdriver or the soldering iron if someones suggests a display overhaul. Thank you!
Regards,
Edgardo Molina
Dirección IPTEL
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
You will be using the counter much on the GPIB than looking to the display,
so try to keep the unit untouched. Better to use it and think about another
one, now that you have one, you are not in a hurry so you can select a
better one. You can also hunt for an SR620 or a Wavecrest... 25pS (or less)
instead of 150pS is better. Isn't it?
On Fri, Nov 2, 2012 at 5:24 PM, Edgardo Molina xe1xus@amsat.org wrote:
Dear Group,
Good morning! I just got my first HP-53132A frequency counter and Prologix
GPIB-USB converter. Now I can do some measurements and I am excited about
it. Everything looks promising. Condition of the instrument on the outside
is like new, controls responsive and no fading in lettering at the buttons.
Connectors are not worn or damaged, original connectors. The unit has been
tested and works OK, probably not with adequate calibration but that is not
an issue now, I can always calibrate it later.
What worries me is the display. Even though being an Agilent built unit,
it has probably more hours on it than a seldom used HP built unit. The
flourescent display is fading, not all the digits show consistent
illumination levels. The left-most digits show some flickering at some of
their segments. It is readable in the end, but I am terribly disappointed
about it.
Does anybody know if there is a cure for this? Or should I start saving
for a better unit.
I am not afraid of the screwdriver or the soldering iron if someones
suggests a display overhaul. Thank you!
Regards,
Edgardo Molina
Dirección IPTEL
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario
de este mensaje, le suplicamos se lo notifique al remitente mediante un
correo electrónico y que borre el presente mensaje y sus anexos de su
computadora sin retener una copia de los mismos. Queda estrictamente
prohibido copiar este mensaje o hacer usode el para cualquier propósito o
divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are
not the intended recipient please immediately advise the sender by replying
to this e-mail and then deleting the message and its attachments from your
computer without keeping a copy. It is strictly forbidden to copy it or use
it for any purpose or disclose its contents to any third party. Thank you.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
The replacement display board is available for about $550. The display itself may be available if you could find the part number for about $125, but Agilent WILL NOT give the raw display part number out. I ran into this same problem with the 3458A and was finally able to find the number and order some. Does anyone on the list have the raw display part number? There is also a small chance the problem is in the HV module. Perhaps going straight to the display vendor.
Thomas Knox
From: xe1xus@amsat.org
Date: Fri, 2 Nov 2012 10:24:07 -0600
To: time-nuts@febo.com
Subject: [time-nuts] HP-53132A Faded and flickering display
Dear Group,
Good morning! I just got my first HP-53132A frequency counter and Prologix GPIB-USB converter. Now I can do some measurements and I am excited about it. Everything looks promising. Condition of the instrument on the outside is like new, controls responsive and no fading in lettering at the buttons. Connectors are not worn or damaged, original connectors. The unit has been tested and works OK, probably not with adequate calibration but that is not an issue now, I can always calibrate it later.
What worries me is the display. Even though being an Agilent built unit, it has probably more hours on it than a seldom used HP built unit. The flourescent display is fading, not all the digits show consistent illumination levels. The left-most digits show some flickering at some of their segments. It is readable in the end, but I am terribly disappointed about it.
Does anybody know if there is a cure for this? Or should I start saving for a better unit.
I am not afraid of the screwdriver or the soldering iron if someones suggests a display overhaul. Thank you!
Regards,
Edgardo Molina
Dirección IPTEL
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Edgardo,
Two possibilities: Cathode poisoning or bad tantalum caps. I'm not sure if "cathode poisoning" is really the correct term for this effect, but it's similar. There have been a number of tutorials out there about rejuvenating VF displays.
The one I have direct experience with is bad caps. We bought an HP 6612C power supply with a bad display for an automated test setup. It's basically the same VF display used in a lot of the instruments from that family.
We didn't really need the display because it was for an automated setup, but I wanted to fix it anyway. When I removed the display board I found that it had 3 or 4 large tantalum caps on one edge. One of the caps was toasted a crispy brown color instead of the usual yellow. After replacing that cap the display came right back to life and has been working fine for a couple of years, now.
BTW: those display and button boards are a bit of a trick to remove. They're slide in from the side under some "L" shaped plastic features in the case front. There are instructions in the service manual for dealing with them.
-John
-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On Behalf Of Edgardo Molina
Sent: Friday, November 02, 2012 11:24 AM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] HP-53132A Faded and flickering display
Dear Group,
Good morning! I just got my first HP-53132A frequency counter and Prologix GPIB-USB converter. Now I can do some measurements and I am excited about it. Everything looks promising. Condition of the instrument on the outside is like new, controls responsive and no fading in lettering at the buttons. Connectors are not worn or damaged, original connectors. The unit has been tested and works OK, probably not with adequate calibration but that is not an issue now, I can always calibrate it later.
What worries me is the display. Even though being an Agilent built unit, it has probably more hours on it than a seldom used HP built unit. The flourescent display is fading, not all the digits show consistent illumination levels. The left-most digits show some flickering at some of their segments. It is readable in the end, but I am terribly disappointed about it.
Does anybody know if there is a cure for this? Or should I start saving for a better unit.
I am not afraid of the screwdriver or the soldering iron if someones suggests a display overhaul. Thank you!
Regards,
Edgardo Molina
Dirección IPTEL
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Dear Azelio,
Thank you. You are right regarding the GPIB operation but since I was looking for a high resolution frequency counter as well for a versatile basic instrument to start measuring with, I choose this frequency counter.
Hunting for a better unit right now will not be a priority as the budget is tight and I can always have one for Christmas without being audited by my significant other ; ) Still, it would be very nice to go down to 25ps for sure!
Thank you for your feedback. Have a great day.
Regards,
Edgardo Molina
Dirección IPTEL
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Nov 2, 2012, at 10:34 AM, Azelio Boriani azelio.boriani@screen.it wrote:
You will be using the counter much on the GPIB than looking to the display,
so try to keep the unit untouched. Better to use it and think about another
one, now that you have one, you are not in a hurry so you can select a
better one. You can also hunt for an SR620 or a Wavecrest... 25pS (or less)
instead of 150pS is better. Isn't it?
On Fri, Nov 2, 2012 at 5:24 PM, Edgardo Molina xe1xus@amsat.org wrote:
Dear Group,
Good morning! I just got my first HP-53132A frequency counter and Prologix
GPIB-USB converter. Now I can do some measurements and I am excited about
it. Everything looks promising. Condition of the instrument on the outside
is like new, controls responsive and no fading in lettering at the buttons.
Connectors are not worn or damaged, original connectors. The unit has been
tested and works OK, probably not with adequate calibration but that is not
an issue now, I can always calibrate it later.
What worries me is the display. Even though being an Agilent built unit,
it has probably more hours on it than a seldom used HP built unit. The
flourescent display is fading, not all the digits show consistent
illumination levels. The left-most digits show some flickering at some of
their segments. It is readable in the end, but I am terribly disappointed
about it.
Does anybody know if there is a cure for this? Or should I start saving
for a better unit.
I am not afraid of the screwdriver or the soldering iron if someones
suggests a display overhaul. Thank you!
Regards,
Edgardo Molina
Dirección IPTEL
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario
de este mensaje, le suplicamos se lo notifique al remitente mediante un
correo electrónico y que borre el presente mensaje y sus anexos de su
computadora sin retener una copia de los mismos. Queda estrictamente
prohibido copiar este mensaje o hacer usode el para cualquier propósito o
divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are
not the intended recipient please immediately advise the sender by replying
to this e-mail and then deleting the message and its attachments from your
computer without keeping a copy. It is strictly forbidden to copy it or use
it for any purpose or disclose its contents to any third party. Thank you.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Unless it is a bad tantalum capacitor, as suggested: in this case yes, it
has to be fixed.
On Fri, Nov 2, 2012 at 5:43 PM, Edgardo Molina xe1xus@amsat.org wrote:
Dear Azelio,
Thank you. You are right regarding the GPIB operation but since I was
looking for a high resolution frequency counter as well for a versatile
basic instrument to start measuring with, I choose this frequency counter.
Hunting for a better unit right now will not be a priority as the budget
is tight and I can always have one for Christmas without being audited by
my significant other ; ) Still, it would be very nice to go down to 25ps
for sure!
Thank you for your feedback. Have a great day.
Regards,
Edgardo Molina
Dirección IPTEL
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario
de este mensaje, le suplicamos se lo notifique al remitente mediante un
correo electrónico y que borre el presente mensaje y sus anexos de su
computadora sin retener una copia de los mismos. Queda estrictamente
prohibido copiar este mensaje o hacer usode el para cualquier propósito o
divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are
not the intended recipient please immediately advise the sender by replying
to this e-mail and then deleting the message and its attachments from your
computer without keeping a copy. It is strictly forbidden to copy it or use
it for any purpose or disclose its contents to any third party. Thank you.
On Nov 2, 2012, at 10:34 AM, Azelio Boriani azelio.boriani@screen.it
wrote:
You will be using the counter much on the GPIB than looking to the
display,
so try to keep the unit untouched. Better to use it and think about
another
one, now that you have one, you are not in a hurry so you can select a
better one. You can also hunt for an SR620 or a Wavecrest... 25pS (or
less)
instead of 150pS is better. Isn't it?
On Fri, Nov 2, 2012 at 5:24 PM, Edgardo Molina xe1xus@amsat.org wrote:
Dear Group,
Good morning! I just got my first HP-53132A frequency counter and
Prologix
GPIB-USB converter. Now I can do some measurements and I am excited
about
it. Everything looks promising. Condition of the instrument on the
outside
is like new, controls responsive and no fading in lettering at the
buttons.
Connectors are not worn or damaged, original connectors. The unit has
been
tested and works OK, probably not with adequate calibration but that is
not
an issue now, I can always calibrate it later.
What worries me is the display. Even though being an Agilent built unit,
it has probably more hours on it than a seldom used HP built unit. The
flourescent display is fading, not all the digits show consistent
illumination levels. The left-most digits show some flickering at some
of
their segments. It is readable in the end, but I am terribly
disappointed
about it.
Does anybody know if there is a cure for this? Or should I start saving
for a better unit.
I am not afraid of the screwdriver or the soldering iron if someones
suggests a display overhaul. Thank you!
Regards,
Edgardo Molina
Dirección IPTEL
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario
de este mensaje, le suplicamos se lo notifique al remitente mediante un
correo electrónico y que borre el presente mensaje y sus anexos de su
computadora sin retener una copia de los mismos. Queda estrictamente
prohibido copiar este mensaje o hacer usode el para cualquier propósito
o
divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you
are
not the intended recipient please immediately advise the sender by
replying
to this e-mail and then deleting the message and its attachments from
your
computer without keeping a copy. It is strictly forbidden to copy it or
use
it for any purpose or disclose its contents to any third party. Thank
you.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
<- snip ->
Two possibilities: Cathode poisoning or bad tantalum caps. I'm not sure if "cathode poisoning" is really the correct term for this effect, but it's similar. There have been a number of tutorials out there about rejuvenating VF displays.
<- snip ->
It appears that the military parts system has the EL displays listed for replacement purposes.
http://www.wbparts.com/rfq/5980-01-396-4645.html
http://www.iso-group.com/Public/PartDetail/CP3033A/12619547/12619547.aspx
Should the display prove to be beyond recovery it might be possible to buy a new one. Now, the cost may be another matter ...
These are made by Noritake-Itron. The generic part number is CP3033A, but it does not show up in their product line on their web site. It might be worth contacting them to see if there's any possibility of still getting them.
http://www.noritake-elec.com/contact_us.htm
Searching for Itron CP3033A turns up some parts brokers that claim to have access to 251 of them.
-John
Hi all,
to avoid more display "consumption" during long acquisitions, there is
an HPIB command to blank the display on the counter. I have asked to
KE5FX to add this feature in Timelab during the acquisition setup. Hope
he have time to do this.
Luciano
Il 2012-11-02 23:13 John Lofgren ha scritto:
<- snip ->
Two possibilities: Cathode poisoning or bad tantalum caps. I'm not
sure if "cathode poisoning" is really the correct term for this
effect, but it's similar. There have been a number of tutorials out
there about rejuvenating VF displays.
<- snip ->
It appears that the military parts system has the EL displays listed
for replacement purposes.
http://www.wbparts.com/rfq/5980-01-396-4645.html
http://www.iso-group.com/Public/PartDetail/CP3033A/12619547/12619547.aspx
Should the display prove to be beyond recovery it might be possible
to buy a new one. Now, the cost may be another matter ...
These are made by Noritake-Itron. The generic part number is
CP3033A, but it does not show up in their product line on their web
site. It might be worth contacting them to see if there's any
possibility of still getting them.
http://www.noritake-elec.com/contact_us.htm
Searching for Itron CP3033A turns up some parts brokers that claim to
have access to 251 of them.
-John
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
--
timeok
Dear Azelio, Tom, John and Luciano ( I hope I am not missing anybody),
Thank you all of you for your kind advise on this subject. I am sorry I was unable to reply in a timely manner. I had to leave town for a week for new IT business. I think I will start by dismantling the unit display and test the tantalum capacitors as you mentioned. If I can achieve success by doing this it is fine. Otherwise I could consider using it as the GPIB workhorse as I just started to obtain my first ADEV, MDEV and TDEV graphs with TimeLab and a Prologix adapter. May I know the command for turning the display off via the GPIB port? That is something nice for long periods of sampling.
A whole new world has just opened to me. Know I am starting to get intimate with my frequency standards by knowing them in deep. If all goes well I should start updating my wish list for a Wavecrest DTS unit next year. For now I will squeeze every single Tau value from the graphs using the 53132A.
Last but not least. If the display fades away I could eventually consider buying a new one as the whole instrument is in pristine condition and would be a shame to see it working without a brilliant display.
Thank you for all your feedback.
Wishing you all a pleasant evening/morning.
Kind regards,
Edgardo Molina
Dirección IPTEL
T : 55 55 55202444
M : 04455 10045822
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Nov 4, 2012, at 8:27 AM, Timeok timeok@timeok.it wrote:
Hi all,
to avoid more display "consumption" during long acquisitions, there is an HPIB command to blank the display on the counter. I have asked to KE5FX to add this feature in Timelab during the acquisition setup. Hope he have time to do this.
Luciano
Il 2012-11-02 23:13 John Lofgren ha scritto:
<- snip ->
Two possibilities: Cathode poisoning or bad tantalum caps. I'm not
sure if "cathode poisoning" is really the correct term for this
effect, but it's similar. There have been a number of tutorials out
there about rejuvenating VF displays.
<- snip ->
It appears that the military parts system has the EL displays listed
for replacement purposes.
http://www.wbparts.com/rfq/5980-01-396-4645.html
http://www.iso-group.com/Public/PartDetail/CP3033A/12619547/12619547.aspx
Should the display prove to be beyond recovery it might be possible
to buy a new one. Now, the cost may be another matter ...
These are made by Noritake-Itron. The generic part number is
CP3033A, but it does not show up in their product line on their web
site. It might be worth contacting them to see if there's any
possibility of still getting them.
http://www.noritake-elec.com/contact_us.htm
Searching for Itron CP3033A turns up some parts brokers that claim to
have access to 251 of them.
-John
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
--
timeok
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Hi Edgardo;
I have been looking for a part number and source. There was a part number posted but it was for the 34401A which is slightly different. I will post price and availability if I find them.
Best Wishes;
Thomas Knox
From: xe1xus@amsat.org
Date: Tue, 13 Nov 2012 21:37:08 -0600
To: time-nuts@febo.com
Subject: Re: [time-nuts] HP-53132A Faded and flickering display
Dear Azelio, Tom, John and Luciano ( I hope I am not missing anybody),
Thank you all of you for your kind advise on this subject. I am sorry I was unable to reply in a timely manner. I had to leave town for a week for new IT business. I think I will start by dismantling the unit display and test the tantalum capacitors as you mentioned. If I can achieve success by doing this it is fine. Otherwise I could consider using it as the GPIB workhorse as I just started to obtain my first ADEV, MDEV and TDEV graphs with TimeLab and a Prologix adapter. May I know the command for turning the display off via the GPIB port? That is something nice for long periods of sampling.
A whole new world has just opened to me. Know I am starting to get intimate with my frequency standards by knowing them in deep. If all goes well I should start updating my wish list for a Wavecrest DTS unit next year. For now I will squeeze every single Tau value from the graphs using the 53132A.
Last but not least. If the display fades away I could eventually consider buying a new one as the whole instrument is in pristine condition and would be a shame to see it working without a brilliant display.
Thank you for all your feedback.
Wishing you all a pleasant evening/morning.
Kind regards,
Edgardo Molina
Dirección IPTEL
T : 55 55 55202444
M : 04455 10045822
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Nov 4, 2012, at 8:27 AM, Timeok timeok@timeok.it wrote:
Hi all,
to avoid more display "consumption" during long acquisitions, there is an HPIB command to blank the display on the counter. I have asked to KE5FX to add this feature in Timelab during the acquisition setup. Hope he have time to do this.
Luciano
Il 2012-11-02 23:13 John Lofgren ha scritto:
<- snip ->
Two possibilities: Cathode poisoning or bad tantalum caps. I'm not
sure if "cathode poisoning" is really the correct term for this
effect, but it's similar. There have been a number of tutorials out
there about rejuvenating VF displays.
<- snip ->
It appears that the military parts system has the EL displays listed
for replacement purposes.
http://www.wbparts.com/rfq/5980-01-396-4645.html
http://www.iso-group.com/Public/PartDetail/CP3033A/12619547/12619547.aspx
Should the display prove to be beyond recovery it might be possible
to buy a new one. Now, the cost may be another matter ...
These are made by Noritake-Itron. The generic part number is
CP3033A, but it does not show up in their product line on their web
site. It might be worth contacting them to see if there's any
possibility of still getting them.
http://www.noritake-elec.com/contact_us.htm
Searching for Itron CP3033A turns up some parts brokers that claim to
have access to 251 of them.
-John
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
--
timeok
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Thanks again Tom! I will wait for your feedback.
Cheers!
Edgardo Molina
Dirección IPTEL
T : 55 55 55202444
M : 04455 10045822
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Nov 13, 2012, at 11:41 PM, Tom Knox actast@hotmail.com wrote:
Hi Edgardo;
I have been looking for a part number and source. There was a part number posted but it was for the 34401A which is slightly different. I will post price and availability if I find them.
Best Wishes;
Thomas Knox
From: xe1xus@amsat.org
Date: Tue, 13 Nov 2012 21:37:08 -0600
To: time-nuts@febo.com
Subject: Re: [time-nuts] HP-53132A Faded and flickering display
Dear Azelio, Tom, John and Luciano ( I hope I am not missing anybody),
Thank you all of you for your kind advise on this subject. I am sorry I was unable to reply in a timely manner. I had to leave town for a week for new IT business. I think I will start by dismantling the unit display and test the tantalum capacitors as you mentioned. If I can achieve success by doing this it is fine. Otherwise I could consider using it as the GPIB workhorse as I just started to obtain my first ADEV, MDEV and TDEV graphs with TimeLab and a Prologix adapter. May I know the command for turning the display off via the GPIB port? That is something nice for long periods of sampling.
A whole new world has just opened to me. Know I am starting to get intimate with my frequency standards by knowing them in deep. If all goes well I should start updating my wish list for a Wavecrest DTS unit next year. For now I will squeeze every single Tau value from the graphs using the 53132A.
Last but not least. If the display fades away I could eventually consider buying a new one as the whole instrument is in pristine condition and would be a shame to see it working without a brilliant display.
Thank you for all your feedback.
Wishing you all a pleasant evening/morning.
Kind regards,
Edgardo Molina
Dirección IPTEL
T : 55 55 55202444
M : 04455 10045822
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Nov 4, 2012, at 8:27 AM, Timeok timeok@timeok.it wrote:
Hi all,
to avoid more display "consumption" during long acquisitions, there is an HPIB command to blank the display on the counter. I have asked to KE5FX to add this feature in Timelab during the acquisition setup. Hope he have time to do this.
Luciano
Il 2012-11-02 23:13 John Lofgren ha scritto:
<- snip ->
Two possibilities: Cathode poisoning or bad tantalum caps. I'm not
sure if "cathode poisoning" is really the correct term for this
effect, but it's similar. There have been a number of tutorials out
there about rejuvenating VF displays.
<- snip ->
It appears that the military parts system has the EL displays listed
for replacement purposes.
http://www.wbparts.com/rfq/5980-01-396-4645.html
http://www.iso-group.com/Public/PartDetail/CP3033A/12619547/12619547.aspx
Should the display prove to be beyond recovery it might be possible
to buy a new one. Now, the cost may be another matter ...
These are made by Noritake-Itron. The generic part number is
CP3033A, but it does not show up in their product line on their web
site. It might be worth contacting them to see if there's any
possibility of still getting them.
http://www.noritake-elec.com/contact_us.htm
Searching for Itron CP3033A turns up some parts brokers that claim to
have access to 251 of them.
-John
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
--
timeok
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Hi Edgardo,all,
the command to blank the display you can find:
http://cp.literature.agilent.com/litweb/pdf/53131-90044.pdf
page 3-36
Disable display:
:DISPlay:ENABle OFF (See Note below.)
Note: This value is not stored on Save and is not part of the
non-volatile state.
Hope this command will be implemented in Timelab soon.
Luciano
Il 2012-11-14 04:37 Edgardo Molina ha scritto:
Dear Azelio, Tom, John and Luciano ( I hope I am not missing
anybody),
Thank you all of you for your kind advise on this subject. I am sorry
I was unable to reply in a timely manner. I had to leave town for a
week for new IT business. I think I will start by dismantling the
unit
display and test the tantalum capacitors as you mentioned. If I can
achieve success by doing this it is fine. Otherwise I could consider
using it as the GPIB workhorse as I just started to obtain my first
ADEV, MDEV and TDEV graphs with TimeLab and a Prologix adapter. May I
know the command for turning the display off via the GPIB port? That
is something nice for long periods of sampling.
A whole new world has just opened to me. Know I am starting to get
intimate with my frequency standards by knowing them in deep. If all
goes well I should start updating my wish list for a Wavecrest DTS
unit next year. For now I will squeeze every single Tau value from
the
graphs using the 53132A.
Last but not least. If the display fades away I could eventually
consider buying a new one as the whole instrument is in pristine
condition and would be a shame to see it working without a brilliant
display.
Thank you for all your feedback.
Wishing you all a pleasant evening/morning.
Kind regards,
Edgardo Molina
Dirección IPTEL
T : 55 55 55202444
M : 04455 10045822
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el
destinarario de este mensaje, le suplicamos se lo notifique al
remitente mediante un correo electrónico y que borre el presente
mensaje y sus anexos de su computadora sin retener una copia de los
mismos. Queda estrictamente prohibido copiar este mensaje o hacer
usode el para cualquier propósito o divulgar su en forma parcial o
total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If
you are not the intended recipient please immediately advise the
sender by replying to this e-mail and then deleting the message and
its attachments from your computer without keeping a copy. It is
strictly forbidden to copy it or use it for any purpose or disclose
its contents to any third party. Thank you.
On Nov 4, 2012, at 8:27 AM, Timeok timeok@timeok.it wrote:
Hi all,
to avoid more display "consumption" during long acquisitions, there
is an HPIB command to blank the display on the counter. I have asked
to KE5FX to add this feature in Timelab during the acquisition setup.
Hope he have time to do this.
Luciano
Il 2012-11-02 23:13 John Lofgren ha scritto:
<- snip ->
Two possibilities: Cathode poisoning or bad tantalum caps. I'm not
sure if "cathode poisoning" is really the correct term for this
effect, but it's similar. There have been a number of tutorials
out
there about rejuvenating VF displays.
<- snip ->
It appears that the military parts system has the EL displays
listed
for replacement purposes.
http://www.wbparts.com/rfq/5980-01-396-4645.html
http://www.iso-group.com/Public/PartDetail/CP3033A/12619547/12619547.aspx
Should the display prove to be beyond recovery it might be possible
to buy a new one. Now, the cost may be another matter ...
These are made by Noritake-Itron. The generic part number is
CP3033A, but it does not show up in their product line on their web
site. It might be worth contacting them to see if there's any
possibility of still getting them.
http://www.noritake-elec.com/contact_us.htm
Searching for Itron CP3033A turns up some parts brokers that claim
to
have access to 251 of them.
-John
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
--
timeok
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
--
timeok
Dear Edgardo,
On 14/11/12 04:37, Edgardo Molina wrote:
Dear Azelio, Tom, John and Luciano ( I hope I am not missing anybody),
Thank you all of you for your kind advise on this subject. I am sorry I was unable to reply in a timely manner. I had to leave town for a week for new IT business. I think I will start by dismantling the unit display and test the tantalum capacitors as you mentioned. If I can achieve success by doing this it is fine. Otherwise I could consider using it as the GPIB workhorse as I just started to obtain my first ADEV, MDEV and TDEV graphs with TimeLab and a Prologix adapter. May I know the command for turning the display off via the GPIB port? That is something nice for long periods of sampling.
A whole new world has just opened to me. Know I am starting to get intimate with my frequency standards by knowing them in deep. If all goes well I should start updating my wish list for a Wavecrest DTS unit next year. For now I will squeeze every single Tau value from the graphs using the 53132A.
Last but not least. If the display fades away I could eventually consider buying a new one as the whole instrument is in pristine condition and would be a shame to see it working without a brilliant display.
Thank you for all your feedback.
Wishing you all a pleasant evening/morning.
It seems like you need to do what I did, go to the agilent site and
download the manuals. Operations and Programming manuals is easy, there
is then a service manual and component-level manual.
The programming manual should assist you on turning of the screen.
The component level manual does not detail the power-supplies.
Did anyone here analyse the analogue interpolators of the 53131/53132?
It's a relative simple design being used with a single transistor and
charging a cap and then read from an ADC. The ADC reference voltages can
be calibrated from a DAC. Notice that there exists two versions of the
main board, for each of the models and that they have different
schematics set in the same manual.
Cheers,
Magnus
Hi
If you go shopping for a parts donor, a 53181 is probably going to be much cheaper. Most of the parts that burn out are common between the models.
Bob
On Nov 13, 2012, at 10:37 PM, Edgardo Molina xe1xus@amsat.org wrote:
Dear Azelio, Tom, John and Luciano ( I hope I am not missing anybody),
Thank you all of you for your kind advise on this subject. I am sorry I was unable to reply in a timely manner. I had to leave town for a week for new IT business. I think I will start by dismantling the unit display and test the tantalum capacitors as you mentioned. If I can achieve success by doing this it is fine. Otherwise I could consider using it as the GPIB workhorse as I just started to obtain my first ADEV, MDEV and TDEV graphs with TimeLab and a Prologix adapter. May I know the command for turning the display off via the GPIB port? That is something nice for long periods of sampling.
A whole new world has just opened to me. Know I am starting to get intimate with my frequency standards by knowing them in deep. If all goes well I should start updating my wish list for a Wavecrest DTS unit next year. For now I will squeeze every single Tau value from the graphs using the 53132A.
Last but not least. If the display fades away I could eventually consider buying a new one as the whole instrument is in pristine condition and would be a shame to see it working without a brilliant display.
Thank you for all your feedback.
Wishing you all a pleasant evening/morning.
Kind regards,
Edgardo Molina
Dirección IPTEL
T : 55 55 55202444
M : 04455 10045822
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Nov 4, 2012, at 8:27 AM, Timeok timeok@timeok.it wrote:
Hi all,
to avoid more display "consumption" during long acquisitions, there is an HPIB command to blank the display on the counter. I have asked to KE5FX to add this feature in Timelab during the acquisition setup. Hope he have time to do this.
Luciano
Il 2012-11-02 23:13 John Lofgren ha scritto:
<- snip ->
Two possibilities: Cathode poisoning or bad tantalum caps. I'm not
sure if "cathode poisoning" is really the correct term for this
effect, but it's similar. There have been a number of tutorials out
there about rejuvenating VF displays.
<- snip ->
It appears that the military parts system has the EL displays listed
for replacement purposes.
http://www.wbparts.com/rfq/5980-01-396-4645.html
http://www.iso-group.com/Public/PartDetail/CP3033A/12619547/12619547.aspx
Should the display prove to be beyond recovery it might be possible
to buy a new one. Now, the cost may be another matter ...
These are made by Noritake-Itron. The generic part number is
CP3033A, but it does not show up in their product line on their web
site. It might be worth contacting them to see if there's any
possibility of still getting them.
http://www.noritake-elec.com/contact_us.htm
Searching for Itron CP3033A turns up some parts brokers that claim to
have access to 251 of them.
-John
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
--
timeok
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Am 14.11.2012 06:41, schrieb Tom Knox:
Hi Edgardo;
I have been looking for a part number and source. There was a part number posted but it was for the 34401A which is slightly different. I will post price and availability if I find them.
Best Wishes;
Thomas Knox
Hi Edgardo,
I bought my used hp-53132A from a local surplus dealer together with a
new spare display, because the display was as well already grey-shaded.
It looks like this is a well known issue on vaccuum fluorescence
displays, because the previous owner was prepared and had some spares in
their stock.
The spare display is manufactured by the japanese company 'itron' and
has the item no. 'CC1187G' on the backside.
I was very lucky to get both together for a reasonable amount.
It does look to be problem now to find such spares, the only company I
found having that part number listed in internet is chip-sale dot com.
(this does not look to be cheap ...)
I was not able to find this number at http://www.noritake-itron.com.
Sorry that I do not have better information, perhaps other groop members
do find a way with this part number.
good luck,
Arnold
Thanks Arnold the more information on this issue the better. I will take a look at your links.
Regards,
Edgardo Molina
Dirección IPTEL
T : 55 55 55202444
M : 04455 10045822
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Nov 14, 2012, at 11:58 AM, Arnold Tibus arnold.tibus@gmx.de wrote:
Am 14.11.2012 06:41, schrieb Tom Knox:
Hi Edgardo;
I have been looking for a part number and source. There was a part number posted but it was for the 34401A which is slightly different. I will post price and availability if I find them.
Best Wishes;
Thomas Knox
Hi Edgardo,
I bought my used hp-53132A from a local surplus dealer together with a
new spare display, because the display was as well already grey-shaded.
It looks like this is a well known issue on vaccuum fluorescence
displays, because the previous owner was prepared and had some spares in
their stock.
The spare display is manufactured by the japanese company 'itron' and
has the item no. 'CC1187G' on the backside.
I was very lucky to get both together for a reasonable amount.
It does look to be problem now to find such spares, the only company I
found having that part number listed in internet is chip-sale dot com.
(this does not look to be cheap ...)
I was not able to find this number at http://www.noritake-itron.com.
Sorry that I do not have better information, perhaps other groop members
do find a way with this part number.
good luck,
Arnold
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
On 11/14/2012 06:58 PM, Arnold Tibus wrote:
Am 14.11.2012 06:41, schrieb Tom Knox:
Hi Edgardo;
I have been looking for a part number and source. There was a part number posted but it was for the 34401A which is slightly different. I will post price and availability if I find them.
Best Wishes;
Thomas Knox
Hi Edgardo,
I bought my used hp-53132A from a local surplus dealer together with a
new spare display, because the display was as well already grey-shaded.
It looks like this is a well known issue on vaccuum fluorescence
displays, because the previous owner was prepared and had some spares in
their stock.
The spare display is manufactured by the japanese company 'itron' and
has the item no. 'CC1187G' on the backside.
The HP manual I have lists DS29 as CP3033A 0950-2327 on the schematic
sheet and PC assembly list as Samsung INB-12MM53T with Agilent P/N
5040-0023.
That should add to the search list of replacements.
Cheers,
Magnus