RN
Ruslan Nabioullin
Thu, Sep 29, 2016 12:07 PM
Hi, I'm unable to fully grasp the lifetime specifications of typical
COTS cesium standard physics packages, for both old and new designs.
For example, the documentation of an older HP standard specifies a shelf
life of two years if the ion pump is operated two or three times per
year, yet at the same time the ``beam tube warranty'' period is three
years. Does this mean that if the physics package is fully operating
(i.e., LOOP OPEN, OPER, or LTC), then it's guaranteed to last at least
three years, but if it's completely shut down, except for vacuum
maintenance sessions (using the mode CS OFF) two or three times per
year, it will last only two years? Also, what will the expected
lifetime of the package be if the standard is kept as a reserve,
specifically, continuously-powered with the mode set to CS OFF (so
continuous vacuum maintenance), and activated or deactivated to or from,
resp., OPER, as needed, excl. the time spent fully operating (i.e., in
this case, the time spent in the OPER mode doesn't count)?
Thanks in advance,
Ruslan
Hi, I'm unable to fully grasp the lifetime specifications of typical
COTS cesium standard physics packages, for both old and new designs.
For example, the documentation of an older HP standard specifies a shelf
life of two years if the ion pump is operated two or three times per
year, yet at the same time the ``beam tube warranty'' period is three
years. Does this mean that if the physics package is fully operating
(i.e., LOOP OPEN, OPER, or LTC), then it's guaranteed to last at least
three years, but if it's completely shut down, except for vacuum
maintenance sessions (using the mode CS OFF) two or three times per
year, it will last only two years? Also, what will the expected
lifetime of the package be if the standard is kept as a reserve,
specifically, continuously-powered with the mode set to CS OFF (so
continuous vacuum maintenance), and activated or deactivated to or from,
resp., OPER, as needed, excl. the time spent fully operating (i.e., in
this case, the time spent in the OPER mode doesn't count)?
Thanks in advance,
Ruslan
SM
Scott McGrath
Thu, Sep 29, 2016 3:45 PM
It depends on the beam tube There is a fixed amount of Cs in any given beam tube this pool of Cs is consumed during operation. Once the Cs is depleted the tube will no longer function. It's possible with intermittent use for the tube to run for decades as long as vacuum maintenance is performed. Most Cs tubes have a published expected lifetime before replacement is required while maintaining published specifications. It's possible to extend tube life somewhat by increasing oven temp and electrode voltages but the signal becomes noisier as a result
Content by Scott
Typos by Siri
On Sep 29, 2016, at 8:07 AM, Ruslan Nabioullin rnabioullin@gmail.com wrote:
Hi, I'm unable to fully grasp the lifetime specifications of typical COTS cesium standard physics packages, for both old and new designs. For example, the documentation of an older HP standard specifies a shelf life of two years if the ion pump is operated two or three times per year, yet at the same time the ``beam tube warranty'' period is three years. Does this mean that if the physics package is fully operating (i.e., LOOP OPEN, OPER, or LTC), then it's guaranteed to last at least three years, but if it's completely shut down, except for vacuum maintenance sessions (using the mode CS OFF) two or three times per year, it will last only two years? Also, what will the expected lifetime of the package be if the standard is kept as a reserve, specifically, continuously-powered with the mode set to CS OFF (so continuous vacuum maintenance), and activated or deactivated to or from, resp., OPER, as needed, excl. the time spent fully operating (i.e., in this case, the time spent in the OPER mode doesn't count)?
Thanks in advance,
Ruslan
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
It depends on the beam tube There is a fixed amount of Cs in any given beam tube this pool of Cs is consumed during operation. Once the Cs is depleted the tube will no longer function. It's possible with intermittent use for the tube to run for decades as long as vacuum maintenance is performed. Most Cs tubes have a published expected lifetime before replacement is required while maintaining published specifications. It's possible to extend tube life somewhat by increasing oven temp and electrode voltages but the signal becomes noisier as a result
Content by Scott
Typos by Siri
> On Sep 29, 2016, at 8:07 AM, Ruslan Nabioullin <rnabioullin@gmail.com> wrote:
>
> Hi, I'm unable to fully grasp the lifetime specifications of typical COTS cesium standard physics packages, for both old and new designs. For example, the documentation of an older HP standard specifies a shelf life of two years if the ion pump is operated two or three times per year, yet at the same time the ``beam tube warranty'' period is three years. Does this mean that if the physics package is fully operating (i.e., LOOP OPEN, OPER, or LTC), then it's guaranteed to last at least three years, but if it's completely shut down, except for vacuum maintenance sessions (using the mode CS OFF) two or three times per year, it will last only two years? Also, what will the expected lifetime of the package be if the standard is kept as a reserve, specifically, continuously-powered with the mode set to CS OFF (so continuous vacuum maintenance), and activated or deactivated to or from, resp., OPER, as needed, excl. the time spent fully operating (i.e., in this case, the time spent in the OPER mode doesn't count)?
>
> Thanks in advance,
> Ruslan
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
BC
Bob Camp
Thu, Sep 29, 2016 5:03 PM
Hi
The real questions are:
-
Over what period of time can you get a replacement under warranty.
-
How long on average is the tube likely to be useful.
There are numbers for each. It just depends on which one you are after.
Bob
On Sep 29, 2016, at 8:07 AM, Ruslan Nabioullin rnabioullin@gmail.com wrote:
Hi, I'm unable to fully grasp the lifetime specifications of typical COTS cesium standard physics packages, for both old and new designs. For example, the documentation of an older HP standard specifies a shelf life of two years if the ion pump is operated two or three times per year, yet at the same time the ``beam tube warranty'' period is three years. Does this mean that if the physics package is fully operating (i.e., LOOP OPEN, OPER, or LTC), then it's guaranteed to last at least three years, but if it's completely shut down, except for vacuum maintenance sessions (using the mode CS OFF) two or three times per year, it will last only two years? Also, what will the expected lifetime of the package be if the standard is kept as a reserve, specifically, continuously-powered with the mode set to CS OFF (so continuous vacuum maintenance), and activated or deactivated to or from, resp., OPER, as needed, excl. the time spent fully operating (i.e., in this case, the time spent in the OPER mode doesn't count)?
Thanks in advance,
Ruslan
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Hi
The real questions are:
1) Over what period of time can you get a replacement under warranty.
2) How long on average is the tube likely to be useful.
There are numbers for each. It just depends on which one you are after.
Bob
> On Sep 29, 2016, at 8:07 AM, Ruslan Nabioullin <rnabioullin@gmail.com> wrote:
>
> Hi, I'm unable to fully grasp the lifetime specifications of typical COTS cesium standard physics packages, for both old and new designs. For example, the documentation of an older HP standard specifies a shelf life of two years if the ion pump is operated two or three times per year, yet at the same time the ``beam tube warranty'' period is three years. Does this mean that if the physics package is fully operating (i.e., LOOP OPEN, OPER, or LTC), then it's guaranteed to last at least three years, but if it's completely shut down, except for vacuum maintenance sessions (using the mode CS OFF) two or three times per year, it will last only two years? Also, what will the expected lifetime of the package be if the standard is kept as a reserve, specifically, continuously-powered with the mode set to CS OFF (so continuous vacuum maintenance), and activated or deactivated to or from, resp., OPER, as needed, excl. the time spent fully operating (i.e., in this case, the time spent in the OPER mode doesn't count)?
>
> Thanks in advance,
> Ruslan
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
RN
Ruslan Nabioullin
Thu, Sep 29, 2016 5:16 PM
On 09/29/2016 01:03 PM, Bob Camp wrote:
- How long on average is the tube likely to be useful.
So you are implying that the package life is depleted at a roughly equal
rate regardless of whether the package is fully operating, has just the
pump operating, or is completely shut off?
-Ruslan
On 09/29/2016 01:03 PM, Bob Camp wrote:
> 2) How long on average is the tube likely to be useful.
So you are implying that the package life is depleted at a roughly equal
rate regardless of whether the package is fully operating, has just the
pump operating, or is completely shut off?
-Ruslan
RN
Ruslan Nabioullin
Thu, Sep 29, 2016 5:46 PM
On 09/29/2016 11:45 AM, Scott McGrath wrote:
It depends on the beam tube There is a fixed amount of Cs in any
given beam tube this pool of Cs is consumed during operation. Once
the Cs is depleted the tube will no longer function. It's possible
with intermittent use for the tube to run for decades as long as
vacuum maintenance is performed. Most Cs tubes have a published
expected lifetime before replacement is required while maintaining
published specifications. It's possible to extend tube life
somewhat by increasing oven temp and electrode voltages but the
signal becomes noisier as a result
Aha, that's what I was thinking---the manual was misleading and made no
sense in this regard. Thanks!
-Ruslan
On 09/29/2016 11:45 AM, Scott McGrath wrote:
> It depends on the beam tube There is a fixed amount of Cs in any
> given beam tube this pool of Cs is consumed during operation. Once
> the Cs is depleted the tube will no longer function. It's possible
> with intermittent use for the tube to run for decades as long as
> vacuum maintenance is performed. Most Cs tubes have a published
> expected lifetime before replacement is required while maintaining
> published specifications. It's possible to extend tube life
> somewhat by increasing oven temp and electrode voltages but the
> signal becomes noisier as a result
Aha, that's what I was thinking---the manual was misleading and made no
sense in this regard. Thanks!
-Ruslan
JA
John Ackermann N8UR
Thu, Sep 29, 2016 5:50 PM
I think there are two separate issues to consider:
-
Cesium depletion, which only occurs when the tube is operating with
cesium oven on, and high voltage at the other end; and
-
Tube vacuum and other physical aspects that may deteriorate over
time, whether or not the tube is operating. The routine ion pumping
helps keep the vacuum up in storage, but is not as effective as the
continuous pumping that occurs during operation.
This is guesswork, but I suspect the warranty period was based on (1),
which is why it is shorter on the high-performance tubes, which deplete
cesium more rapidly. There may also have been an assumption that not
many customers were buying Cs standards and putting them in storage.
John
On 9/29/2016 1:16 PM, Ruslan Nabioullin wrote:
On 09/29/2016 01:03 PM, Bob Camp wrote:
- How long on average is the tube likely to be useful.
I think there are two separate issues to consider:
1. Cesium depletion, which only occurs when the tube is operating with
cesium oven on, and high voltage at the other end; and
2. Tube vacuum and other physical aspects that may deteriorate over
time, whether or not the tube is operating. The routine ion pumping
helps keep the vacuum up in storage, but is not as effective as the
continuous pumping that occurs during operation.
This is guesswork, but I suspect the warranty period was based on (1),
which is why it is shorter on the high-performance tubes, which deplete
cesium more rapidly. There may also have been an assumption that not
many customers were buying Cs standards and putting them in storage.
John
----
On 9/29/2016 1:16 PM, Ruslan Nabioullin wrote:
> On 09/29/2016 01:03 PM, Bob Camp wrote:
>> 2) How long on average is the tube likely to be useful.
>
> So you are implying that the package life is depleted at a roughly equal
> rate regardless of whether the package is fully operating, has just the
> pump operating, or is completely shut off?
>
> -Ruslan
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
RN
Ruslan Nabioullin
Thu, Sep 29, 2016 5:56 PM
On 09/29/2016 01:50 PM, John Ackermann N8UR wrote:
- Tube vacuum and other physical aspects that may deteriorate over
time, whether or not the tube is operating. The routine ion pumping
helps keep the vacuum up in storage, but is not as effective as the
continuous pumping that occurs during operation.
This is guesswork, but I suspect the warranty period was based on (1),
which is why it is shorter on the high-performance tubes, which deplete
cesium more rapidly. There may also have been an assumption that not
many customers were buying Cs standards and putting them in storage.
But I'm interested in a different case---having the unit be continuously
in operation (so continuous pumping, not two or three times per the
period), but having the cesium physics package be in rare, intermittent
operation.
-Ruslan
On 09/29/2016 01:50 PM, John Ackermann N8UR wrote:
> 2. Tube vacuum and other physical aspects that may deteriorate over
> time, whether or not the tube is operating. The routine ion pumping
> helps keep the vacuum up in storage, but is not as effective as the
> continuous pumping that occurs during operation.
>
> This is guesswork, but I suspect the warranty period was based on (1),
> which is why it is shorter on the high-performance tubes, which deplete
> cesium more rapidly. There may also have been an assumption that not
> many customers were buying Cs standards and putting them in storage.
But I'm interested in a different case---having the unit be continuously
in operation (so continuous pumping, not two or three times per the
period), but having the cesium physics package be in rare, intermittent
operation.
-Ruslan
BC
Bob Camp
Thu, Sep 29, 2016 5:59 PM
On Sep 29, 2016, at 1:16 PM, Ruslan Nabioullin rnabioullin@gmail.com wrote:
On 09/29/2016 01:03 PM, Bob Camp wrote:
- How long on average is the tube likely to be useful.
So you are implying that the package life is depleted at a roughly equal rate regardless of whether the package is fully operating, has just the pump operating, or is completely shut off?
Not really. What I am suggesting is that the warranty covers a variety of issues. The practical use of the part is normally something fairly specific. You might leave the unit in the box and never power it up. You might also run it 24 hours a day 7 days a week. Both are pretty common. In the 24 / 7 run case, vacuum is not an issue at all. It’s entirely Cs depletion and that is fairly predictable. In the “never runs” case there are a lot of variables involving seal leak rates, part outgassing and getter capacity. Two of the three are pretty much a guess. That makes the “power off” case much less predictable.
Simply put:
It is rare to find a high performance tube that runs for much over 10 years. It is not uncommon to find a tube that has been on the shelf for a decade or two that works.
Bob
Hi
> On Sep 29, 2016, at 1:16 PM, Ruslan Nabioullin <rnabioullin@gmail.com> wrote:
>
> On 09/29/2016 01:03 PM, Bob Camp wrote:
>> 2) How long on average is the tube likely to be useful.
>
> So you are implying that the package life is depleted at a roughly equal rate regardless of whether the package is fully operating, has just the pump operating, or is completely shut off?
Not really. What I am suggesting is that the warranty covers a variety of issues. The practical use of the part is normally something fairly specific. You might leave the unit in the box and never power it up. You might also run it 24 hours a day 7 days a week. Both are pretty common. In the 24 / 7 run case, vacuum is not an issue at all. It’s entirely Cs depletion and that is fairly predictable. In the “never runs” case there are a lot of variables involving seal leak rates, part outgassing and getter capacity. Two of the three are pretty much a guess. That makes the “power off” case much less predictable.
Simply put:
It is rare to find a high performance tube that runs for much over 10 years. It is not uncommon to find a tube that has been on the shelf for a decade or two that works.
Bob
>
> -Ruslan
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
BC
Bob Camp
Thu, Sep 29, 2016 6:02 PM
Hi
Are you interested in improving the likelihood of the unit working 30 years out or of it
working 15 years out? The “operating” case will ultimately kill various parts in the standard
that would not die sitting on the shelf. The pumping process may extend the life of the tube…..
What the pumping process will do is reduce the amount of time needed to bring the unit
back to life after it has been on the shelf for 30 years.
Bob
On Sep 29, 2016, at 1:56 PM, Ruslan Nabioullin rnabioullin@gmail.com wrote:
On 09/29/2016 01:50 PM, John Ackermann N8UR wrote:
- Tube vacuum and other physical aspects that may deteriorate over
time, whether or not the tube is operating. The routine ion pumping
helps keep the vacuum up in storage, but is not as effective as the
continuous pumping that occurs during operation.
This is guesswork, but I suspect the warranty period was based on (1),
which is why it is shorter on the high-performance tubes, which deplete
cesium more rapidly. There may also have been an assumption that not
many customers were buying Cs standards and putting them in storage.
But I'm interested in a different case---having the unit be continuously in operation (so continuous pumping, not two or three times per the period), but having the cesium physics package be in rare, intermittent operation.
-Ruslan
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Hi
Are you interested in improving the likelihood of the unit working 30 years out or of it
working 15 years out? The “operating” case will ultimately kill various parts in the standard
that would not die sitting on the shelf. The pumping process *may* extend the life of the tube…..
What the pumping process *will* do is reduce the amount of time needed to bring the unit
back to life after it has been on the shelf for 30 years.
Bob
> On Sep 29, 2016, at 1:56 PM, Ruslan Nabioullin <rnabioullin@gmail.com> wrote:
>
> On 09/29/2016 01:50 PM, John Ackermann N8UR wrote:
>> 2. Tube vacuum and other physical aspects that may deteriorate over
>> time, whether or not the tube is operating. The routine ion pumping
>> helps keep the vacuum up in storage, but is not as effective as the
>> continuous pumping that occurs during operation.
>>
>> This is guesswork, but I suspect the warranty period was based on (1),
>> which is why it is shorter on the high-performance tubes, which deplete
>> cesium more rapidly. There may also have been an assumption that not
>> many customers were buying Cs standards and putting them in storage.
>
> But I'm interested in a different case---having the unit be continuously in operation (so continuous pumping, not two or three times per the period), but having the cesium physics package be in rare, intermittent operation.
>
> -Ruslan
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
RN
Ruslan Nabioullin
Thu, Sep 29, 2016 6:08 PM
On 09/29/2016 02:02 PM, Bob Camp wrote:
The “operating” case will ultimately kill various parts in the
standard that would not die sitting on the shelf.
Oh, really? I thought that cesium depletion is essentially the only
real killer.
-Ruslan
On 09/29/2016 02:02 PM, Bob Camp wrote:
> The “operating” case will ultimately kill various parts in the
> standard that would not die sitting on the shelf.
Oh, really? I thought that cesium depletion is essentially the only
real killer.
-Ruslan
PS
paul swed
Thu, Sep 29, 2016 6:16 PM
The tubes were never intended to last 10 and more years. Bad for business.
The reality as time-nuts have discovered is that they do. What you have in
this group is an amazing knowledge of how to keep them ticking. Pun
intended.
The comment of raising the oven temp is exactly how I was able to get my
5061 going again.I took this somewhat to an extreme and the crazy thing
works and relocks even today. But here I suspect that the downside of
raising the temp is that it may also shift the frequency or transparency as
a better statement.
Talk about in the noise. Really can't see current.
Regards
Paul
WB8TSL
On Thu, Sep 29, 2016 at 1:46 PM, Ruslan Nabioullin rnabioullin@gmail.com
wrote:
On 09/29/2016 11:45 AM, Scott McGrath wrote:
It depends on the beam tube There is a fixed amount of Cs in any
given beam tube this pool of Cs is consumed during operation. Once
the Cs is depleted the tube will no longer function. It's possible
with intermittent use for the tube to run for decades as long as
vacuum maintenance is performed. Most Cs tubes have a published
expected lifetime before replacement is required while maintaining
published specifications. It's possible to extend tube life
somewhat by increasing oven temp and electrode voltages but the
signal becomes noisier as a result
Aha, that's what I was thinking---the manual was misleading and made no
sense in this regard. Thanks!
-Ruslan
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.
The tubes were never intended to last 10 and more years. Bad for business.
The reality as time-nuts have discovered is that they do. What you have in
this group is an amazing knowledge of how to keep them ticking. Pun
intended.
The comment of raising the oven temp is exactly how I was able to get my
5061 going again.I took this somewhat to an extreme and the crazy thing
works and relocks even today. But here I suspect that the downside of
raising the temp is that it may also shift the frequency or transparency as
a better statement.
Talk about in the noise. Really can't see current.
Regards
Paul
WB8TSL
On Thu, Sep 29, 2016 at 1:46 PM, Ruslan Nabioullin <rnabioullin@gmail.com>
wrote:
> On 09/29/2016 11:45 AM, Scott McGrath wrote:
>
>> It depends on the beam tube There is a fixed amount of Cs in any
>> given beam tube this pool of Cs is consumed during operation. Once
>> the Cs is depleted the tube will no longer function. It's possible
>> with intermittent use for the tube to run for decades as long as
>> vacuum maintenance is performed. Most Cs tubes have a published
>> expected lifetime before replacement is required while maintaining
>> published specifications. It's possible to extend tube life
>> somewhat by increasing oven temp and electrode voltages but the
>> signal becomes noisier as a result
>>
>
> Aha, that's what I was thinking---the manual was misleading and made no
> sense in this regard. Thanks!
>
>
> -Ruslan
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/m
> ailman/listinfo/time-nuts
> and follow the instructions there.
>
BC
Bob Camp
Thu, Sep 29, 2016 6:39 PM
Hi
If you operate the device continuously for 30 years, you will have other parts in
it fail. The MTBF of the system (without the tube) is not infinite.
Bob
On Sep 29, 2016, at 2:08 PM, Ruslan Nabioullin rnabioullin@gmail.com wrote:
On 09/29/2016 02:02 PM, Bob Camp wrote:
The “operating” case will ultimately kill various parts in the
standard that would not die sitting on the shelf.
Hi
If you operate the device continuously for 30 years, you *will* have other parts in
it fail. The MTBF of the system (without the tube) is *not* infinite.
Bob
> On Sep 29, 2016, at 2:08 PM, Ruslan Nabioullin <rnabioullin@gmail.com> wrote:
>
> On 09/29/2016 02:02 PM, Bob Camp wrote:
>> The “operating” case will ultimately kill various parts in the
>> standard that would not die sitting on the shelf.
>
> Oh, really? I thought that cesium depletion is essentially the only real killer.
>
> -Ruslan
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
JM
John Miles
Thu, Sep 29, 2016 10:05 PM
- Cesium depletion, which only occurs when the tube is operating with
cesium oven on, and high voltage at the other end; and
The Cs atoms need to be electrically neutral, so their depletion rate shouldn't depend on the presence of HV, just the oven heater. State selection wouldn't work on ionized atoms, and you also wouldn't want them to be accelerated towards the electron multiplier until after they've reached the ionizer filament at the end of the cavity. (The longer they spend hanging out with Schroedinger's cat in the Ramsey cavity, the narrower the line width.)
- Tube vacuum and other physical aspects that may deteriorate over
time, whether or not the tube is operating. The routine ion pumping
helps keep the vacuum up in storage, but is not as effective as the
continuous pumping that occurs during operation.
Some of the HP standards (5060, 5061, 5071) have a CS OFF setting that enables the ion pump by itself. Those tubes should last indefinitely in that mode of operation. Unfortunately the 5062C didn't have a way to disable the Cs oven while leaving the ion pump active, which is probably why there are so few operational 5062C tubes left. By now, most are either out of cesium or too gassy for the ion pump to recover.
I added a switch to my 5062C to allow the vacuum to be maintained without running the Cs oven. It gets turned on every so often, maybe a couple of times per year, when I want a noisy signal source with known ADEV characteristics. Even though the ion pump runs 24/7/365, there are always a few overcurrent cycles on initial powerup for some reason, where the beam current exceeds the trip point and shuts down the power supplies. Something inside the tube is apparently outgassing during warmup -- whether it's the Cs oven or the hot-wire ionizer ribbon, I don't know. But the condition always clears itself within a few seconds. The same thing happens when I turn my 5061s on after a long period in CS OFF.
- john, KE5FX
Miles Design LLC
> 1. Cesium depletion, which only occurs when the tube is operating with
> cesium oven on, and high voltage at the other end; and
The Cs atoms need to be electrically neutral, so their depletion rate shouldn't depend on the presence of HV, just the oven heater. State selection wouldn't work on ionized atoms, and you also wouldn't want them to be accelerated towards the electron multiplier until after they've reached the ionizer filament at the end of the cavity. (The longer they spend hanging out with Schroedinger's cat in the Ramsey cavity, the narrower the line width.)
> 2. Tube vacuum and other physical aspects that may deteriorate over
> time, whether or not the tube is operating. The routine ion pumping
> helps keep the vacuum up in storage, but is not as effective as the
> continuous pumping that occurs during operation.
Some of the HP standards (5060, 5061, 5071) have a CS OFF setting that enables the ion pump by itself. Those tubes should last indefinitely in that mode of operation. Unfortunately the 5062C didn't have a way to disable the Cs oven while leaving the ion pump active, which is probably why there are so few operational 5062C tubes left. By now, most are either out of cesium or too gassy for the ion pump to recover.
I added a switch to my 5062C to allow the vacuum to be maintained without running the Cs oven. It gets turned on every so often, maybe a couple of times per year, when I want a noisy signal source with known ADEV characteristics. Even though the ion pump runs 24/7/365, there are always a few overcurrent cycles on initial powerup for some reason, where the beam current exceeds the trip point and shuts down the power supplies. Something inside the tube is apparently outgassing during warmup -- whether it's the Cs oven or the hot-wire ionizer ribbon, I don't know. But the condition always clears itself within a few seconds. The same thing happens when I turn my 5061s on after a long period in CS OFF.
- john, KE5FX
Miles Design LLC
TM
Tom Miller
Thu, Sep 29, 2016 10:37 PM
Hi John,
I have a Datum FTS-5060 unit. Did you just open up the heater line to the
beam tube? I like the idea of conserving the Cs when pumping down the tube
for maintenance.
Regards,
Tom
----- Original Message -----
From: "John Miles" john@miles.io
To: "'Discussion of precise time and frequency measurement'"
time-nuts@febo.com
Sent: Thursday, September 29, 2016 6:05 PM
Subject: Re: [time-nuts] COTS cesium standard physics package life
- Cesium depletion, which only occurs when the tube is operating with
cesium oven on, and high voltage at the other end; and
The Cs atoms need to be electrically neutral, so their depletion rate
shouldn't depend on the presence of HV, just the oven heater. State
selection wouldn't work on ionized atoms, and you also wouldn't want them
to be accelerated towards the electron multiplier until after they've
reached the ionizer filament at the end of the cavity. (The longer they
spend hanging out with Schroedinger's cat in the Ramsey cavity, the
narrower the line width.)
- Tube vacuum and other physical aspects that may deteriorate over
time, whether or not the tube is operating. The routine ion pumping
helps keep the vacuum up in storage, but is not as effective as the
continuous pumping that occurs during operation.
Some of the HP standards (5060, 5061, 5071) have a CS OFF setting that
enables the ion pump by itself. Those tubes should last indefinitely in
that mode of operation. Unfortunately the 5062C didn't have a way to
disable the Cs oven while leaving the ion pump active, which is probably
why there are so few operational 5062C tubes left. By now, most are
either out of cesium or too gassy for the ion pump to recover.
I added a switch to my 5062C to allow the vacuum to be maintained without
running the Cs oven. It gets turned on every so often, maybe a couple of
times per year, when I want a noisy signal source with known ADEV
characteristics. Even though the ion pump runs 24/7/365, there are always
a few overcurrent cycles on initial powerup for some reason, where the
beam current exceeds the trip point and shuts down the power supplies.
Something inside the tube is apparently outgassing during warmup --
whether it's the Cs oven or the hot-wire ionizer ribbon, I don't know.
But the condition always clears itself within a few seconds. The same
thing happens when I turn my 5061s on after a long period in CS OFF.
- john, KE5FX
Miles Design LLC
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Hi John,
I have a Datum FTS-5060 unit. Did you just open up the heater line to the
beam tube? I like the idea of conserving the Cs when pumping down the tube
for maintenance.
Regards,
Tom
----- Original Message -----
From: "John Miles" <john@miles.io>
To: "'Discussion of precise time and frequency measurement'"
<time-nuts@febo.com>
Sent: Thursday, September 29, 2016 6:05 PM
Subject: Re: [time-nuts] COTS cesium standard physics package life
>> 1. Cesium depletion, which only occurs when the tube is operating with
>> cesium oven on, and high voltage at the other end; and
>
> The Cs atoms need to be electrically neutral, so their depletion rate
> shouldn't depend on the presence of HV, just the oven heater. State
> selection wouldn't work on ionized atoms, and you also wouldn't want them
> to be accelerated towards the electron multiplier until after they've
> reached the ionizer filament at the end of the cavity. (The longer they
> spend hanging out with Schroedinger's cat in the Ramsey cavity, the
> narrower the line width.)
>
>> 2. Tube vacuum and other physical aspects that may deteriorate over
>> time, whether or not the tube is operating. The routine ion pumping
>> helps keep the vacuum up in storage, but is not as effective as the
>> continuous pumping that occurs during operation.
>
> Some of the HP standards (5060, 5061, 5071) have a CS OFF setting that
> enables the ion pump by itself. Those tubes should last indefinitely in
> that mode of operation. Unfortunately the 5062C didn't have a way to
> disable the Cs oven while leaving the ion pump active, which is probably
> why there are so few operational 5062C tubes left. By now, most are
> either out of cesium or too gassy for the ion pump to recover.
>
> I added a switch to my 5062C to allow the vacuum to be maintained without
> running the Cs oven. It gets turned on every so often, maybe a couple of
> times per year, when I want a noisy signal source with known ADEV
> characteristics. Even though the ion pump runs 24/7/365, there are always
> a few overcurrent cycles on initial powerup for some reason, where the
> beam current exceeds the trip point and shuts down the power supplies.
> Something inside the tube is apparently outgassing during warmup --
> whether it's the Cs oven or the hot-wire ionizer ribbon, I don't know.
> But the condition always clears itself within a few seconds. The same
> thing happens when I turn my 5061s on after a long period in CS OFF.
>
> - john, KE5FX
> Miles Design LLC
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
JM
John Miles
Thu, Sep 29, 2016 11:44 PM
If I remember correctly, the oven power supply runs off of the same bus that powers the rest of the unit, in the 5062C's case. It took a fair bit of hacking to install the switch and make it do what I wanted. Not much besides the ion pump and the OCXO remains active with the switch in the 'off' position.
I'd expect the Datum unit to be pretty different in terms of its schematic and power distribution scheme, but I'd shoot for the same functionality if I were installing a switch in one.
-- john, KE5FX
Miles Design LLC
I have a Datum FTS-5060 unit. Did you just open up the heater line to the
beam tube? I like the idea of conserving the Cs when pumping down the
tube
for maintenance.
Regards,
Tom
If I remember correctly, the oven power supply runs off of the same bus that powers the rest of the unit, in the 5062C's case. It took a fair bit of hacking to install the switch and make it do what I wanted. Not much besides the ion pump and the OCXO remains active with the switch in the 'off' position.
I'd expect the Datum unit to be pretty different in terms of its schematic and power distribution scheme, but I'd shoot for the same functionality if I were installing a switch in one.
-- john, KE5FX
Miles Design LLC
> I have a Datum FTS-5060 unit. Did you just open up the heater line to the
> beam tube? I like the idea of conserving the Cs when pumping down the
> tube
> for maintenance.
>
> Regards,
> Tom
>