I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to lock
and produce a nice sine wave that stays in phase with my Trimble Tbolt for
a time which exceeds my attention span!!!
It does however get hot in operation. As this is my first experience with a
Rb standard I don’t know whether this is normal or indicates that I need a
heat sink.
Any comments would be appreciated.
Regards
Richard Katsch
VK2EIK
Hello All,
I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to lock
and produce a nice sine wave that stays in phase with my Trimble Tbolt for
a time which exceeds my attention span!!!
It does however get hot in operation. As this is my first experience with a
Rb standard I don’t know whether this is normal or indicates that I need a
heat sink.
Any comments would be appreciated.
Regards
Richard Katsch
VK2EIK
MW
Matthias Welwarsky
Tue, Sep 1, 2020 7:01 AM
On Dienstag, 1. September 2020 07:15:34 CEST Richard Katsch wrote:
Hello All,
I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to lock
and produce a nice sine wave that stays in phase with my Trimble Tbolt for
a time which exceeds my attention span!!!
It does however get hot in operation. As this is my first experience with a
Rb standard I don’t know whether this is normal or indicates that I need a
heat sink.
Any comments would be appreciated.
They're supposed to run hot, but excess temperature will degrade the stability
and shorten the lifetime of the electronics. It's hard to recommend what
heatsink they require if you have no documentation. Typically you'll need one.
My LPRO-101 need a heatsink with < 2K/W of thermal resistance to keep the
temperature in check.
BR,
Matthias
On Dienstag, 1. September 2020 07:15:34 CEST Richard Katsch wrote:
> Hello All,
> I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to lock
> and produce a nice sine wave that stays in phase with my Trimble Tbolt for
> a time which exceeds my attention span!!!
> It does however get hot in operation. As this is my first experience with a
> Rb standard I don’t know whether this is normal or indicates that I need a
> heat sink.
> Any comments would be appreciated.
They're supposed to run hot, but excess temperature will degrade the stability
and shorten the lifetime of the electronics. It's hard to recommend what
heatsink they require if you have no documentation. Typically you'll need one.
My LPRO-101 need a heatsink with < 2K/W of thermal resistance to keep the
temperature in check.
BR,
Matthias
> Regards
> Richard Katsch
> VK2EIK
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow
> the instructions there.
HB
Hugh Blemings
Tue, Sep 1, 2020 9:02 AM
Hiya,
I'd wondered this also - my (limited) experience of Rbs are the LPROs
and they definitely need a heatsink (I have a nice story to share about
getting some made one day which I must write up, but I digress)
Curious, I dug around a bit and only data I can find is a two pager on
the manufacturer website -
https://www.quartzlock.com/product/Rubidium/rubidium-oscillators/E10-MRX
The packaging itself doesn't appear to be designed with heatsinking in
mind, but perhaps there is an implicit expectation of convection cooling
at least to keep it manageable ?
Any idea what case temperature you're seeing @Richard ?
Cheers,
Hugh
On 1/9/20 5:01 pm, Matthias Welwarsky wrote:
On Dienstag, 1. September 2020 07:15:34 CEST Richard Katsch wrote:
Hello All,
I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to lock
and produce a nice sine wave that stays in phase with my Trimble Tbolt for
a time which exceeds my attention span!!!
It does however get hot in operation. As this is my first experience with a
Rb standard I don’t know whether this is normal or indicates that I need a
heat sink.
Any comments would be appreciated.
They're supposed to run hot, but excess temperature will degrade the stability
and shorten the lifetime of the electronics. It's hard to recommend what
heatsink they require if you have no documentation. Typically you'll need one.
My LPRO-101 need a heatsink with < 2K/W of thermal resistance to keep the
temperature in check.
BR,
Matthias
Hiya,
I'd wondered this also - my (limited) experience of Rbs are the LPROs
and they definitely need a heatsink (I have a nice story to share about
getting some made one day which I must write up, but I digress)
Curious, I dug around a bit and only data I can find is a two pager on
the manufacturer website -
https://www.quartzlock.com/product/Rubidium/rubidium-oscillators/E10-MRX
The packaging itself doesn't appear to be designed with heatsinking in
mind, but perhaps there is an implicit expectation of convection cooling
at least to keep it manageable ?
Any idea what case temperature you're seeing @Richard ?
Cheers,
Hugh
On 1/9/20 5:01 pm, Matthias Welwarsky wrote:
> On Dienstag, 1. September 2020 07:15:34 CEST Richard Katsch wrote:
>> Hello All,
>> I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to lock
>> and produce a nice sine wave that stays in phase with my Trimble Tbolt for
>> a time which exceeds my attention span!!!
>> It does however get hot in operation. As this is my first experience with a
>> Rb standard I don’t know whether this is normal or indicates that I need a
>> heat sink.
>> Any comments would be appreciated.
> They're supposed to run hot, but excess temperature will degrade the stability
> and shorten the lifetime of the electronics. It's hard to recommend what
> heatsink they require if you have no documentation. Typically you'll need one.
> My LPRO-101 need a heatsink with < 2K/W of thermal resistance to keep the
> temperature in check.
>
> BR,
> Matthias
>
>> Regards
>> Richard Katsch
>> VK2EIK
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@lists.febo.com
>> To unsubscribe, go to
>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow
>> the instructions there.
>
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.
RK
Richard Katsch
Tue, Sep 1, 2020 12:25 PM
Many thanks Matthias,Hugh.
The case doesn’t seem to have any direct method of fastening to a heatsink
but something can be arranged I’m sure. The manufacturer’s documents don’t
suggest anything. They do say that the base temperature can be up to 85
degrees C which seems awfully hot.
I haven’t measured the temperature other than a finger test that indicates
that more than 5 secs is uncomfortably hot. At that point I switched it off.
Best regards
Richard
On Tue, 1 Sep 2020 at 19:02, Hugh Blemings hugh@blemings.org wrote:
Hiya,
I'd wondered this also - my (limited) experience of Rbs are the
LPROs and they definitely need a heatsink (I have a nice story to
share about getting some made one day which I must write up, but I
digress)
Curious, I dug around a bit and only data I can find is a two
pager on the manufacturer website -
https://www.quartzlock.com/product/Rubidium/rubidium-oscillators/E10-MRX
The packaging itself doesn't appear to be designed with
heatsinking in mind, but perhaps there is an implicit expectation
of convection cooling at least to keep it manageable ?
Any idea what case temperature you're seeing @Richard ?
Cheers,
Hugh
On 1/9/20 5:01 pm, Matthias Welwarsky
wrote:
On Dienstag, 1. September 2020 07:15:34 CEST Richard Katsch wrote:
Hello All,
I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to lock
and produce a nice sine wave that stays in phase with my Trimble Tbolt for
a time which exceeds my attention span!!!
It does however get hot in operation. As this is my first experience with a
Rb standard I don’t know whether this is normal or indicates that I need a
heat sink.
Any comments would be appreciated.
They're supposed to run hot, but excess temperature will degrade the stability
and shorten the lifetime of the electronics. It's hard to recommend what
heatsink they require if you have no documentation. Typically you'll need one.
My LPRO-101 need a heatsink with < 2K/W of thermal resistance to keep the
temperature in check.
BR,
Matthias
Regards
Richard Katsch
VK2EIK
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow
the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Many thanks Matthias,Hugh.
The case doesn’t seem to have any direct method of fastening to a heatsink
but something can be arranged I’m sure. The manufacturer’s documents don’t
suggest anything. They do say that the base temperature can be up to 85
degrees C which seems awfully hot.
I haven’t measured the temperature other than a finger test that indicates
that more than 5 secs is uncomfortably hot. At that point I switched it off.
Best regards
Richard
On Tue, 1 Sep 2020 at 19:02, Hugh Blemings <hugh@blemings.org> wrote:
>
>
>
>
>
>
>
>
>
>
> Hiya,
>
>
> I'd wondered this also - my (limited) experience of Rbs are the
>
> LPROs and they definitely need a heatsink (I have a nice story to
>
> share about getting some made one day which I must write up, but I
>
> digress)
>
>
> Curious, I dug around a bit and only data I can find is a two
>
> pager on the manufacturer website -
>
> https://www.quartzlock.com/product/Rubidium/rubidium-oscillators/E10-MRX
>
>
> The packaging itself doesn't appear to be designed with
>
> heatsinking in mind, but perhaps there is an implicit expectation
>
> of convection cooling at least to keep it manageable ?
>
>
> Any idea what case temperature you're seeing @Richard ?
>
>
>
>
> Cheers,
>
>
> Hugh
>
>
>
>
>
>
>
> On 1/9/20 5:01 pm, Matthias Welwarsky
>
> wrote:
>
>
>
>
>
>
> On Dienstag, 1. September 2020 07:15:34 CEST Richard Katsch wrote:
>
>
>
>
>
> Hello All,
>
> I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to lock
>
> and produce a nice sine wave that stays in phase with my Trimble Tbolt for
>
> a time which exceeds my attention span!!!
>
> It does however get hot in operation. As this is my first experience with a
>
> Rb standard I don’t know whether this is normal or indicates that I need a
>
> heat sink.
>
> Any comments would be appreciated.
>
>
>
>
>
>
> They're supposed to run hot, but excess temperature will degrade the stability
>
> and shorten the lifetime of the electronics. It's hard to recommend what
>
> heatsink they require if you have no documentation. Typically you'll need one.
>
> My LPRO-101 need a heatsink with < 2K/W of thermal resistance to keep the
>
> temperature in check.
>
>
>
> BR,
>
> Matthias
>
>
>
>
>
>
>
> Regards
>
> Richard Katsch
>
> VK2EIK
>
> _______________________________________________
>
> time-nuts mailing list -- time-nuts@lists.febo.com
>
> To unsubscribe, go to
>
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow
>
> the instructions there.
>
>
>
>
>
>
>
>
>
>
>
>
>
>
> _______________________________________________
>
> time-nuts mailing list -- time-nuts@lists.febo.com
>
> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
>
> and follow the instructions there.
>
>
>
>
>
>
>
>
>
>
MW
Matthias Welwarsky
Tue, Sep 1, 2020 1:23 PM
On Dienstag, 1. September 2020 14:25:31 CEST Richard Katsch wrote:
Many thanks Matthias,Hugh.
The case doesn’t seem to have any direct method of fastening to a heatsink
but something can be arranged I’m sure. The manufacturer’s documents don’t
suggest anything. They do say that the base temperature can be up to 85
degrees C which seems awfully hot.
I haven’t measured the temperature other than a finger test that indicates
that more than 5 secs is uncomfortably hot. At that point I switched it off.
I wouldn't say "uncomfortably hot" is a cause for alarm. The lamp housing in
the LPRO-101 runs at 100°C, the resonance chamber at roughly the same temp.
You need Rb vapor at around that temperature (if that Rb is a traditional
construction) for the system to work.
Anyway, what else can you do? If it's broken, the chances of repairing it are
slim. If it works, no way to find out if it's in spec unless you run it. Just
watch for smoke ;)
BR,
Matthias
Best regards
Richard
On Tue, 1 Sep 2020 at 19:02, Hugh Blemings hugh@blemings.org wrote:
Hiya,
I'd wondered this also - my (limited) experience of Rbs are the
LPROs and they definitely need a heatsink (I have a nice story to
share about getting some made one day which I must write up, but I
digress)
Curious, I dug around a bit and only data I can find is a two
pager on the manufacturer website -
https://www.quartzlock.com/product/Rubidium/rubidium-oscillators/E10-MRX
The packaging itself doesn't appear to be designed with
heatsinking in mind, but perhaps there is an implicit expectation
of convection cooling at least to keep it manageable ?
Any idea what case temperature you're seeing @Richard ?
Cheers,
Hugh
On 1/9/20 5:01 pm, Matthias Welwarsky
wrote:
On Dienstag, 1. September 2020 07:15:34 CEST Richard Katsch wrote:
Hello All,
I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to
lock
and produce a nice sine wave that stays in phase with my Trimble Tbolt for
a time which exceeds my attention span!!!
It does however get hot in operation. As this is my first experience with
a
Rb standard I don’t know whether this is normal or indicates that I need a
heat sink.
Any comments would be appreciated.
They're supposed to run hot, but excess temperature will degrade the
stability
and shorten the lifetime of the electronics. It's hard to recommend what
heatsink they require if you have no documentation. Typically you'll need
one.
My LPRO-101 need a heatsink with < 2K/W of thermal resistance to keep the
temperature in check.
BR,
Matthias
Regards
Richard Katsch
VK2EIK
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow
the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
On Dienstag, 1. September 2020 14:25:31 CEST Richard Katsch wrote:
> Many thanks Matthias,Hugh.
>
> The case doesn’t seem to have any direct method of fastening to a heatsink
> but something can be arranged I’m sure. The manufacturer’s documents don’t
> suggest anything. They do say that the base temperature can be up to 85
> degrees C which seems awfully hot.
>
> I haven’t measured the temperature other than a finger test that indicates
> that more than 5 secs is uncomfortably hot. At that point I switched it off.
I wouldn't say "uncomfortably hot" is a cause for alarm. The lamp housing in
the LPRO-101 runs at 100°C, the resonance chamber at roughly the same temp.
You need Rb vapor at around that temperature (if that Rb is a traditional
construction) for the system to work.
Anyway, what else can you do? If it's broken, the chances of repairing it are
slim. If it works, no way to find out if it's in spec unless you run it. Just
watch for smoke ;)
BR,
Matthias
>
> Best regards
> Richard
>
> On Tue, 1 Sep 2020 at 19:02, Hugh Blemings <hugh@blemings.org> wrote:
> > Hiya,
> >
> >
> > I'd wondered this also - my (limited) experience of Rbs are the
> >
> > LPROs and they definitely need a heatsink (I have a nice story to
> >
> > share about getting some made one day which I must write up, but I
> >
> > digress)
> >
> >
> > Curious, I dug around a bit and only data I can find is a two
> >
> > pager on the manufacturer website -
> >
> > https://www.quartzlock.com/product/Rubidium/rubidium-oscillators/E10-MRX
> >
> >
> > The packaging itself doesn't appear to be designed with
> >
> > heatsinking in mind, but perhaps there is an implicit expectation
> >
> > of convection cooling at least to keep it manageable ?
> >
> >
> > Any idea what case temperature you're seeing @Richard ?
> >
> >
> >
> >
> > Cheers,
> >
> >
> > Hugh
> >
> >
> >
> >
> >
> >
> >
> > On 1/9/20 5:01 pm, Matthias Welwarsky
> >
> > wrote:
> >
> >
> >
> >
> >
> >
> > On Dienstag, 1. September 2020 07:15:34 CEST Richard Katsch wrote:
> >
> >
> >
> >
> >
> > Hello All,
> >
> > I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to
> > lock
> >
> > and produce a nice sine wave that stays in phase with my Trimble Tbolt for
> >
> > a time which exceeds my attention span!!!
> >
> > It does however get hot in operation. As this is my first experience with
> > a
> >
> > Rb standard I don’t know whether this is normal or indicates that I need a
> >
> > heat sink.
> >
> > Any comments would be appreciated.
> >
> >
> >
> >
> >
> >
> > They're supposed to run hot, but excess temperature will degrade the
> > stability
> >
> > and shorten the lifetime of the electronics. It's hard to recommend what
> >
> > heatsink they require if you have no documentation. Typically you'll need
> > one.
> >
> > My LPRO-101 need a heatsink with < 2K/W of thermal resistance to keep the
> >
> > temperature in check.
> >
> >
> >
> > BR,
> >
> > Matthias
> >
> >
> >
> >
> >
> >
> >
> > Regards
> >
> > Richard Katsch
> >
> > VK2EIK
> >
> > _______________________________________________
> >
> > time-nuts mailing list -- time-nuts@lists.febo.com
> >
> > To unsubscribe, go to
> >
> > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow
> >
> > the instructions there.
> >
> >
> >
> >
> >
> >
> >
> >
> >
> >
> >
> >
> >
> >
> > _______________________________________________
> >
> > time-nuts mailing list -- time-nuts@lists.febo.com
> >
> > To unsubscribe, go to
> > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> >
> > and follow the instructions there.
BK
Bob kb8tq
Tue, Sep 1, 2020 2:07 PM
Hi
In this case there may not be much that can be done. In the more general case
( = a unit with a surface that mates to a heatsink) , it’s well worth doing something.
I have an unfortunate lot of empirical evidence (on FRK’s and LPRO’s ) showing that
a “no heatsink” setup is one with a short life. With no heatsink, a couple of years is
doing well on a surplus unit. With a heatsink, > 10 years is quite possible.
How much heatsink? Is the next question. If you can get the baseplate down into the
40’s (C) that’s doing pretty well. A fan plus heatsink may be needed to get this done.
If a fan is involved, mag field could become an issue ( = Rb’s are sensitive to mag
field, as the fan speed changes the field may change).
=====
Indeed this is not always the official word from the Rb manufacturers. I’ve spent
considerable effort drinking beer with many of their marketing and engineering
guys. The conversation is not quite the same when marketing it not at the table.
I believe that they simply don’t have a body of long term data “in house” to study.
The fact that I’m not the only one seeing Rb’s die early is backed up on eBay.
From time to time, you can find pretty large batches of “for parts” Rb’s. Date codes
often are not very far back in time.
If they get cheap enough (like ~ $30) they can be fun to play with. You can pretty
quickly work out that the electrolytic caps are the most likely failure points. Cap
swap outs may only fix 2/3 of them, but it’s quick and easy. You now have a pile
of $45 Rb’s ….
Bob
On Sep 1, 2020, at 9:23 AM, Matthias Welwarsky time-nuts@welwarsky.de wrote:
On Dienstag, 1. September 2020 14:25:31 CEST Richard Katsch wrote:
Many thanks Matthias,Hugh.
The case doesn’t seem to have any direct method of fastening to a heatsink
but something can be arranged I’m sure. The manufacturer’s documents don’t
suggest anything. They do say that the base temperature can be up to 85
degrees C which seems awfully hot.
I haven’t measured the temperature other than a finger test that indicates
that more than 5 secs is uncomfortably hot. At that point I switched it off.
I wouldn't say "uncomfortably hot" is a cause for alarm. The lamp housing in
the LPRO-101 runs at 100°C, the resonance chamber at roughly the same temp.
You need Rb vapor at around that temperature (if that Rb is a traditional
construction) for the system to work.
Anyway, what else can you do? If it's broken, the chances of repairing it are
slim. If it works, no way to find out if it's in spec unless you run it. Just
watch for smoke ;)
BR,
Matthias
Best regards
Richard
On Tue, 1 Sep 2020 at 19:02, Hugh Blemings hugh@blemings.org wrote:
Hiya,
I'd wondered this also - my (limited) experience of Rbs are the
LPROs and they definitely need a heatsink (I have a nice story to
share about getting some made one day which I must write up, but I
digress)
Curious, I dug around a bit and only data I can find is a two
pager on the manufacturer website -
https://www.quartzlock.com/product/Rubidium/rubidium-oscillators/E10-MRX
The packaging itself doesn't appear to be designed with
heatsinking in mind, but perhaps there is an implicit expectation
of convection cooling at least to keep it manageable ?
Any idea what case temperature you're seeing @Richard ?
Cheers,
Hugh
On 1/9/20 5:01 pm, Matthias Welwarsky
wrote:
On Dienstag, 1. September 2020 07:15:34 CEST Richard Katsch wrote:
Hello All,
I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to
lock
and produce a nice sine wave that stays in phase with my Trimble Tbolt for
a time which exceeds my attention span!!!
It does however get hot in operation. As this is my first experience with
a
Rb standard I don’t know whether this is normal or indicates that I need a
heat sink.
Any comments would be appreciated.
They're supposed to run hot, but excess temperature will degrade the
stability
and shorten the lifetime of the electronics. It's hard to recommend what
heatsink they require if you have no documentation. Typically you'll need
one.
My LPRO-101 need a heatsink with < 2K/W of thermal resistance to keep the
temperature in check.
BR,
Matthias
Regards
Richard Katsch
VK2EIK
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow
the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Hi
In this case there may not be much that can be done. In the more general case
( = a unit with a surface that mates to a heatsink) , it’s well worth doing something.
I have an unfortunate lot of empirical evidence (on FRK’s and LPRO’s ) showing that
a “no heatsink” setup is one with a short life. With no heatsink, a couple of years is
doing well on a surplus unit. With a heatsink, > 10 years is quite possible.
How much heatsink? Is the next question. If you can get the baseplate down into the
40’s (C) that’s doing pretty well. A fan plus heatsink may be needed to get this done.
If a fan is involved, mag field could become an issue ( = Rb’s are sensitive to mag
field, as the fan speed changes the field *may* change).
=====
Indeed this is not always the official word from the Rb manufacturers. I’ve spent
considerable effort drinking beer with many of their marketing and engineering
guys. The conversation is not *quite* the same when marketing it not at the table.
I believe that they simply don’t have a body of long term data “in house” to study.
The fact that I’m not the only one seeing Rb’s die early is backed up on eBay.
From time to time, you can find pretty large batches of “for parts” Rb’s. Date codes
often are *not* very far back in time.
If they get cheap enough (like ~ $30) they can be fun to play with. You can pretty
quickly work out that the electrolytic caps are the most likely failure points. Cap
swap outs may only fix 2/3 of them, but it’s quick and easy. You now have a pile
of $45 Rb’s ….
Bob
> On Sep 1, 2020, at 9:23 AM, Matthias Welwarsky <time-nuts@welwarsky.de> wrote:
>
> On Dienstag, 1. September 2020 14:25:31 CEST Richard Katsch wrote:
>> Many thanks Matthias,Hugh.
>>
>> The case doesn’t seem to have any direct method of fastening to a heatsink
>> but something can be arranged I’m sure. The manufacturer’s documents don’t
>> suggest anything. They do say that the base temperature can be up to 85
>> degrees C which seems awfully hot.
>>
>> I haven’t measured the temperature other than a finger test that indicates
>> that more than 5 secs is uncomfortably hot. At that point I switched it off.
>
> I wouldn't say "uncomfortably hot" is a cause for alarm. The lamp housing in
> the LPRO-101 runs at 100°C, the resonance chamber at roughly the same temp.
> You need Rb vapor at around that temperature (if that Rb is a traditional
> construction) for the system to work.
>
> Anyway, what else can you do? If it's broken, the chances of repairing it are
> slim. If it works, no way to find out if it's in spec unless you run it. Just
> watch for smoke ;)
>
> BR,
> Matthias
>
>>
>> Best regards
>> Richard
>>
>> On Tue, 1 Sep 2020 at 19:02, Hugh Blemings <hugh@blemings.org> wrote:
>>> Hiya,
>>>
>>>
>>> I'd wondered this also - my (limited) experience of Rbs are the
>>>
>>> LPROs and they definitely need a heatsink (I have a nice story to
>>>
>>> share about getting some made one day which I must write up, but I
>>>
>>> digress)
>>>
>>>
>>> Curious, I dug around a bit and only data I can find is a two
>>>
>>> pager on the manufacturer website -
>>>
>>> https://www.quartzlock.com/product/Rubidium/rubidium-oscillators/E10-MRX
>>>
>>>
>>> The packaging itself doesn't appear to be designed with
>>>
>>> heatsinking in mind, but perhaps there is an implicit expectation
>>>
>>> of convection cooling at least to keep it manageable ?
>>>
>>>
>>> Any idea what case temperature you're seeing @Richard ?
>>>
>>>
>>>
>>>
>>> Cheers,
>>>
>>>
>>> Hugh
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>> On 1/9/20 5:01 pm, Matthias Welwarsky
>>>
>>> wrote:
>>>
>>>
>>>
>>>
>>>
>>>
>>> On Dienstag, 1. September 2020 07:15:34 CEST Richard Katsch wrote:
>>>
>>>
>>>
>>>
>>>
>>> Hello All,
>>>
>>> I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to
>>> lock
>>>
>>> and produce a nice sine wave that stays in phase with my Trimble Tbolt for
>>>
>>> a time which exceeds my attention span!!!
>>>
>>> It does however get hot in operation. As this is my first experience with
>>> a
>>>
>>> Rb standard I don’t know whether this is normal or indicates that I need a
>>>
>>> heat sink.
>>>
>>> Any comments would be appreciated.
>>>
>>>
>>>
>>>
>>>
>>>
>>> They're supposed to run hot, but excess temperature will degrade the
>>> stability
>>>
>>> and shorten the lifetime of the electronics. It's hard to recommend what
>>>
>>> heatsink they require if you have no documentation. Typically you'll need
>>> one.
>>>
>>> My LPRO-101 need a heatsink with < 2K/W of thermal resistance to keep the
>>>
>>> temperature in check.
>>>
>>>
>>>
>>> BR,
>>>
>>> Matthias
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>> Regards
>>>
>>> Richard Katsch
>>>
>>> VK2EIK
>>>
>>> _______________________________________________
>>>
>>> time-nuts mailing list -- time-nuts@lists.febo.com
>>>
>>> To unsubscribe, go to
>>>
>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow
>>>
>>> the instructions there.
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>> _______________________________________________
>>>
>>> time-nuts mailing list -- time-nuts@lists.febo.com
>>>
>>> To unsubscribe, go to
>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
>>>
>>> and follow the instructions there.
>
>
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.
TK
Taka Kamiya
Tue, Sep 1, 2020 8:32 PM
Will you elaborate on fan? Just about everything I've read, use of fan was discouraged. Natural convection over heat sink fins were recommended. I've done limited testing. When fan cooling, indeed temperature will go down, but current consumption stays higher indicating heating is often conducted to keep it hot inside. How do I reconciliate these opposing conditions in actual use?
(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG
On Tuesday, September 1, 2020, 2:08:04 PM EDT, Bob kb8tq <kb8tq@n1k.org> wrote:
Hi
In this case there may not be much that can be done. In the more general case
( = a unit with a surface that mates to a heatsink) , it’s well worth doing something.
I have an unfortunate lot of empirical evidence (on FRK’s and LPRO’s ) showing that
a “no heatsink” setup is one with a short life. With no heatsink, a couple of years is
doing well on a surplus unit. With a heatsink, > 10 years is quite possible.
How much heatsink? Is the next question. If you can get the baseplate down into the
40’s (C) that’s doing pretty well. A fan plus heatsink may be needed to get this done.
If a fan is involved, mag field could become an issue ( = Rb’s are sensitive to mag
field, as the fan speed changes the field may change).
=====
Indeed this is not always the official word from the Rb manufacturers. I’ve spent
considerable effort drinking beer with many of their marketing and engineering
guys. The conversation is not quite the same when marketing it not at the table.
I believe that they simply don’t have a body of long term data “in house” to study.
The fact that I’m not the only one seeing Rb’s die early is backed up on eBay.
From time to time, you can find pretty large batches of “for parts” Rb’s. Date codes
often are not very far back in time.
If they get cheap enough (like ~ $30) they can be fun to play with. You can pretty
quickly work out that the electrolytic caps are the most likely failure points. Cap
swap outs may only fix 2/3 of them, but it’s quick and easy. You now have a pile
of $45 Rb’s ….
Bob
On Sep 1, 2020, at 9:23 AM, Matthias Welwarsky time-nuts@welwarsky.de wrote:
On Dienstag, 1. September 2020 14:25:31 CEST Richard Katsch wrote:
Many thanks Matthias,Hugh.
The case doesn’t seem to have any direct method of fastening to a heatsink
but something can be arranged I’m sure. The manufacturer’s documents don’t
suggest anything. They do say that the base temperature can be up to 85
degrees C which seems awfully hot.
I haven’t measured the temperature other than a finger test that indicates
that more than 5 secs is uncomfortably hot. At that point I switched it off.
I wouldn't say "uncomfortably hot" is a cause for alarm. The lamp housing in
the LPRO-101 runs at 100°C, the resonance chamber at roughly the same temp.
You need Rb vapor at around that temperature (if that Rb is a traditional
construction) for the system to work.
Anyway, what else can you do? If it's broken, the chances of repairing it are
slim. If it works, no way to find out if it's in spec unless you run it. Just
watch for smoke ;)
BR,
Matthias
Best regards
Richard
On Tue, 1 Sep 2020 at 19:02, Hugh Blemings hugh@blemings.org wrote:
Hiya,
I'd wondered this also - my (limited) experience of Rbs are the
LPROs and they definitely need a heatsink (I have a nice story to
share about getting some made one day which I must write up, but I
digress)
Curious, I dug around a bit and only data I can find is a two
pager on the manufacturer website -
https://www.quartzlock.com/product/Rubidium/rubidium-oscillators/E10-MRX
The packaging itself doesn't appear to be designed with
heatsinking in mind, but perhaps there is an implicit expectation
of convection cooling at least to keep it manageable ?
Any idea what case temperature you're seeing @Richard ?
Cheers,
Hugh
On 1/9/20 5:01 pm, Matthias Welwarsky
wrote:
On Dienstag, 1. September 2020 07:15:34 CEST Richard Katsch wrote:
Hello All,
I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to
lock
and produce a nice sine wave that stays in phase with my Trimble Tbolt for
a time which exceeds my attention span!!!
It does however get hot in operation. As this is my first experience with
a
Rb standard I don’t know whether this is normal or indicates that I need a
heat sink.
Any comments would be appreciated.
They're supposed to run hot, but excess temperature will degrade the
stability
and shorten the lifetime of the electronics. It's hard to recommend what
heatsink they require if you have no documentation. Typically you'll need
one.
My LPRO-101 need a heatsink with < 2K/W of thermal resistance to keep the
temperature in check.
BR,
Matthias
Regards
Richard Katsch
VK2EIK
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow
the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Will you elaborate on fan? Just about everything I've read, use of fan was discouraged. Natural convection over heat sink fins were recommended. I've done limited testing. When fan cooling, indeed temperature will go down, but current consumption stays higher indicating heating is often conducted to keep it hot inside. How do I reconciliate these opposing conditions in actual use?
---------------------------------------
(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG
On Tuesday, September 1, 2020, 2:08:04 PM EDT, Bob kb8tq <kb8tq@n1k.org> wrote:
Hi
In this case there may not be much that can be done. In the more general case
( = a unit with a surface that mates to a heatsink) , it’s well worth doing something.
I have an unfortunate lot of empirical evidence (on FRK’s and LPRO’s ) showing that
a “no heatsink” setup is one with a short life. With no heatsink, a couple of years is
doing well on a surplus unit. With a heatsink, > 10 years is quite possible.
How much heatsink? Is the next question. If you can get the baseplate down into the
40’s (C) that’s doing pretty well. A fan plus heatsink may be needed to get this done.
If a fan is involved, mag field could become an issue ( = Rb’s are sensitive to mag
field, as the fan speed changes the field *may* change).
=====
Indeed this is not always the official word from the Rb manufacturers. I’ve spent
considerable effort drinking beer with many of their marketing and engineering
guys. The conversation is not *quite* the same when marketing it not at the table.
I believe that they simply don’t have a body of long term data “in house” to study.
The fact that I’m not the only one seeing Rb’s die early is backed up on eBay.
>From time to time, you can find pretty large batches of “for parts” Rb’s. Date codes
often are *not* very far back in time.
If they get cheap enough (like ~ $30) they can be fun to play with. You can pretty
quickly work out that the electrolytic caps are the most likely failure points. Cap
swap outs may only fix 2/3 of them, but it’s quick and easy. You now have a pile
of $45 Rb’s ….
Bob
> On Sep 1, 2020, at 9:23 AM, Matthias Welwarsky <time-nuts@welwarsky.de> wrote:
>
> On Dienstag, 1. September 2020 14:25:31 CEST Richard Katsch wrote:
>> Many thanks Matthias,Hugh.
>>
>> The case doesn’t seem to have any direct method of fastening to a heatsink
>> but something can be arranged I’m sure. The manufacturer’s documents don’t
>> suggest anything. They do say that the base temperature can be up to 85
>> degrees C which seems awfully hot.
>>
>> I haven’t measured the temperature other than a finger test that indicates
>> that more than 5 secs is uncomfortably hot. At that point I switched it off.
>
> I wouldn't say "uncomfortably hot" is a cause for alarm. The lamp housing in
> the LPRO-101 runs at 100°C, the resonance chamber at roughly the same temp.
> You need Rb vapor at around that temperature (if that Rb is a traditional
> construction) for the system to work.
>
> Anyway, what else can you do? If it's broken, the chances of repairing it are
> slim. If it works, no way to find out if it's in spec unless you run it. Just
> watch for smoke ;)
>
> BR,
> Matthias
>
>>
>> Best regards
>> Richard
>>
>> On Tue, 1 Sep 2020 at 19:02, Hugh Blemings <hugh@blemings.org> wrote:
>>> Hiya,
>>>
>>>
>>> I'd wondered this also - my (limited) experience of Rbs are the
>>>
>>> LPROs and they definitely need a heatsink (I have a nice story to
>>>
>>> share about getting some made one day which I must write up, but I
>>>
>>> digress)
>>>
>>>
>>> Curious, I dug around a bit and only data I can find is a two
>>>
>>> pager on the manufacturer website -
>>>
>>> https://www.quartzlock.com/product/Rubidium/rubidium-oscillators/E10-MRX
>>>
>>>
>>> The packaging itself doesn't appear to be designed with
>>>
>>> heatsinking in mind, but perhaps there is an implicit expectation
>>>
>>> of convection cooling at least to keep it manageable ?
>>>
>>>
>>> Any idea what case temperature you're seeing @Richard ?
>>>
>>>
>>>
>>>
>>> Cheers,
>>>
>>>
>>> Hugh
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>> On 1/9/20 5:01 pm, Matthias Welwarsky
>>>
>>> wrote:
>>>
>>>
>>>
>>>
>>>
>>>
>>> On Dienstag, 1. September 2020 07:15:34 CEST Richard Katsch wrote:
>>>
>>>
>>>
>>>
>>>
>>> Hello All,
>>>
>>> I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to
>>> lock
>>>
>>> and produce a nice sine wave that stays in phase with my Trimble Tbolt for
>>>
>>> a time which exceeds my attention span!!!
>>>
>>> It does however get hot in operation. As this is my first experience with
>>> a
>>>
>>> Rb standard I don’t know whether this is normal or indicates that I need a
>>>
>>> heat sink.
>>>
>>> Any comments would be appreciated.
>>>
>>>
>>>
>>>
>>>
>>>
>>> They're supposed to run hot, but excess temperature will degrade the
>>> stability
>>>
>>> and shorten the lifetime of the electronics. It's hard to recommend what
>>>
>>> heatsink they require if you have no documentation. Typically you'll need
>>> one.
>>>
>>> My LPRO-101 need a heatsink with < 2K/W of thermal resistance to keep the
>>>
>>> temperature in check.
>>>
>>>
>>>
>>> BR,
>>>
>>> Matthias
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>> Regards
>>>
>>> Richard Katsch
>>>
>>> VK2EIK
>>>
>>> _______________________________________________
>>>
>>> time-nuts mailing list -- time-nuts@lists.febo.com
>>>
>>> To unsubscribe, go to
>>>
>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow
>>>
>>> the instructions there.
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>> _______________________________________________
>>>
>>> time-nuts mailing list -- time-nuts@lists.febo.com
>>>
>>> To unsubscribe, go to
>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
>>>
>>> and follow the instructions there.
>
>
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.
_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
R
rfnuts
Tue, Sep 1, 2020 9:52 PM
If it works, don't try to fix it.
In the datasheet
https://quartzlock.com/userfiles/downloads/datasheets/e10-mrx(2)(1).pdf
there is no cooling requirement specified.
I would assume that it is usually mounted on a PCB which adds some
cooling though not terribly much.
After 1 hour the case temperature should be about 45 Deg. C at 25 Deg.
ambient.
Adrian
Am 01.09.2020 um 07:15 schrieb Richard Katsch:
Hello All,
I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to lock
and produce a nice sine wave that stays in phase with my Trimble Tbolt for
a time which exceeds my attention span!!!
It does however get hot in operation. As this is my first experience with a
Rb standard I don’t know whether this is normal or indicates that I need a
heat sink.
Any comments would be appreciated.
Regards
Richard Katsch
VK2EIK
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
If it works, don't try to fix it.
In the datasheet
https://quartzlock.com/userfiles/downloads/datasheets/e10-mrx(2)(1).pdf
there is no cooling requirement specified.
I would assume that it is usually mounted on a PCB which adds some
cooling though not terribly much.
After 1 hour the case temperature should be about 45 Deg. C at 25 Deg.
ambient.
Adrian
Am 01.09.2020 um 07:15 schrieb Richard Katsch:
> Hello All,
> I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to lock
> and produce a nice sine wave that stays in phase with my Trimble Tbolt for
> a time which exceeds my attention span!!!
> It does however get hot in operation. As this is my first experience with a
> Rb standard I don’t know whether this is normal or indicates that I need a
> heat sink.
> Any comments would be appreciated.
> Regards
> Richard Katsch
> VK2EIK
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.
>
BK
Bob kb8tq
Tue, Sep 1, 2020 10:23 PM
Hi
If you have enough room for a heatsink that is the size of a couple of shoe boxes, then
by all means use a convection heatsink.
There is zero difference between cooling by fan vs big heatsink as far as device current
consumption is concerned. This or that level of cooling takes the baseplate to this or that
temperature. At that baseplate temperature, the device pulls X watts (regardless of how
you get to that temperature).
If your “natural convection” approach does not increase the current drain then it’s not
actually cooling the baseplate in any meaningful way. The current drain goes up to heat
the oven. The baseplate cools to protect the rest of the device.
Indeed the fan its self pulls some current. That current is typically quite low compared to
the power you are trying to manage. One cute thing with a fan is that you can servo it.
Baseplate control of well under 1C is fairly easy to do.
Fun !!
Bob
On Sep 1, 2020, at 4:32 PM, Taka Kamiya via time-nuts time-nuts@lists.febo.com wrote:
Will you elaborate on fan? Just about everything I've read, use of fan was discouraged. Natural convection over heat sink fins were recommended. I've done limited testing. When fan cooling, indeed temperature will go down, but current consumption stays higher indicating heating is often conducted to keep it hot inside. How do I reconciliate these opposing conditions in actual use?
(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG
On Tuesday, September 1, 2020, 2:08:04 PM EDT, Bob kb8tq <kb8tq@n1k.org> wrote:
Hi
In this case there may not be much that can be done. In the more general case
( = a unit with a surface that mates to a heatsink) , it’s well worth doing something.
I have an unfortunate lot of empirical evidence (on FRK’s and LPRO’s ) showing that
a “no heatsink” setup is one with a short life. With no heatsink, a couple of years is
doing well on a surplus unit. With a heatsink, > 10 years is quite possible.
How much heatsink? Is the next question. If you can get the baseplate down into the
40’s (C) that’s doing pretty well. A fan plus heatsink may be needed to get this done.
If a fan is involved, mag field could become an issue ( = Rb’s are sensitive to mag
field, as the fan speed changes the field may change).
=====
Indeed this is not always the official word from the Rb manufacturers. I’ve spent
considerable effort drinking beer with many of their marketing and engineering
guys. The conversation is not quite the same when marketing it not at the table.
I believe that they simply don’t have a body of long term data “in house” to study.
The fact that I’m not the only one seeing Rb’s die early is backed up on eBay.
From time to time, you can find pretty large batches of “for parts” Rb’s. Date codes
often are not very far back in time.
If they get cheap enough (like ~ $30) they can be fun to play with. You can pretty
quickly work out that the electrolytic caps are the most likely failure points. Cap
swap outs may only fix 2/3 of them, but it’s quick and easy. You now have a pile
of $45 Rb’s ….
Bob
On Sep 1, 2020, at 9:23 AM, Matthias Welwarsky time-nuts@welwarsky.de wrote:
On Dienstag, 1. September 2020 14:25:31 CEST Richard Katsch wrote:
Many thanks Matthias,Hugh.
The case doesn’t seem to have any direct method of fastening to a heatsink
but something can be arranged I’m sure. The manufacturer’s documents don’t
suggest anything. They do say that the base temperature can be up to 85
degrees C which seems awfully hot.
I haven’t measured the temperature other than a finger test that indicates
that more than 5 secs is uncomfortably hot. At that point I switched it off.
I wouldn't say "uncomfortably hot" is a cause for alarm. The lamp housing in
the LPRO-101 runs at 100°C, the resonance chamber at roughly the same temp.
You need Rb vapor at around that temperature (if that Rb is a traditional
construction) for the system to work.
Anyway, what else can you do? If it's broken, the chances of repairing it are
slim. If it works, no way to find out if it's in spec unless you run it. Just
watch for smoke ;)
BR,
Matthias
Best regards
Richard
On Tue, 1 Sep 2020 at 19:02, Hugh Blemings hugh@blemings.org wrote:
Hiya,
I'd wondered this also - my (limited) experience of Rbs are the
LPROs and they definitely need a heatsink (I have a nice story to
share about getting some made one day which I must write up, but I
digress)
Curious, I dug around a bit and only data I can find is a two
pager on the manufacturer website -
https://www.quartzlock.com/product/Rubidium/rubidium-oscillators/E10-MRX
The packaging itself doesn't appear to be designed with
heatsinking in mind, but perhaps there is an implicit expectation
of convection cooling at least to keep it manageable ?
Any idea what case temperature you're seeing @Richard ?
Cheers,
Hugh
On 1/9/20 5:01 pm, Matthias Welwarsky
wrote:
On Dienstag, 1. September 2020 07:15:34 CEST Richard Katsch wrote:
Hello All,
I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to
lock
and produce a nice sine wave that stays in phase with my Trimble Tbolt for
a time which exceeds my attention span!!!
It does however get hot in operation. As this is my first experience with
a
Rb standard I don’t know whether this is normal or indicates that I need a
heat sink.
Any comments would be appreciated.
They're supposed to run hot, but excess temperature will degrade the
stability
and shorten the lifetime of the electronics. It's hard to recommend what
heatsink they require if you have no documentation. Typically you'll need
one.
My LPRO-101 need a heatsink with < 2K/W of thermal resistance to keep the
temperature in check.
BR,
Matthias
Regards
Richard Katsch
VK2EIK
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow
the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Hi
If you have enough room for a heatsink that is the size of a couple of shoe boxes, then
by all means use a convection heatsink.
There is zero difference between cooling by fan vs big heatsink as far as device current
consumption is concerned. This or that level of cooling takes the baseplate to this or that
temperature. At that baseplate temperature, the device pulls X watts (regardless of how
you get to that temperature).
If your “natural convection” approach does *not* increase the current drain then it’s not
actually cooling the baseplate in any meaningful way. The current drain goes up to heat
the oven. The baseplate cools to protect the rest of the device.
Indeed the fan its self pulls some current. That current is typically quite low compared to
the power you are trying to manage. One cute thing with a fan is that you can servo it.
Baseplate control of well under 1C is fairly easy to do.
Fun !!
Bob
> On Sep 1, 2020, at 4:32 PM, Taka Kamiya via time-nuts <time-nuts@lists.febo.com> wrote:
>
> Will you elaborate on fan? Just about everything I've read, use of fan was discouraged. Natural convection over heat sink fins were recommended. I've done limited testing. When fan cooling, indeed temperature will go down, but current consumption stays higher indicating heating is often conducted to keep it hot inside. How do I reconciliate these opposing conditions in actual use?
>
> ---------------------------------------
> (Mr.) Taka Kamiya
> KB4EMF / ex JF2DKG
>
>
> On Tuesday, September 1, 2020, 2:08:04 PM EDT, Bob kb8tq <kb8tq@n1k.org> wrote:
>
> Hi
>
> In this case there may not be much that can be done. In the more general case
> ( = a unit with a surface that mates to a heatsink) , it’s well worth doing something.
>
> I have an unfortunate lot of empirical evidence (on FRK’s and LPRO’s ) showing that
> a “no heatsink” setup is one with a short life. With no heatsink, a couple of years is
> doing well on a surplus unit. With a heatsink, > 10 years is quite possible.
>
> How much heatsink? Is the next question. If you can get the baseplate down into the
> 40’s (C) that’s doing pretty well. A fan plus heatsink may be needed to get this done.
> If a fan is involved, mag field could become an issue ( = Rb’s are sensitive to mag
> field, as the fan speed changes the field *may* change).
>
> =====
>
> Indeed this is not always the official word from the Rb manufacturers. I’ve spent
> considerable effort drinking beer with many of their marketing and engineering
> guys. The conversation is not *quite* the same when marketing it not at the table.
> I believe that they simply don’t have a body of long term data “in house” to study.
>
> The fact that I’m not the only one seeing Rb’s die early is backed up on eBay.
> From time to time, you can find pretty large batches of “for parts” Rb’s. Date codes
> often are *not* very far back in time.
>
> If they get cheap enough (like ~ $30) they can be fun to play with. You can pretty
> quickly work out that the electrolytic caps are the most likely failure points. Cap
> swap outs may only fix 2/3 of them, but it’s quick and easy. You now have a pile
> of $45 Rb’s ….
>
> Bob
>
>
>
>
>> On Sep 1, 2020, at 9:23 AM, Matthias Welwarsky <time-nuts@welwarsky.de> wrote:
>>
>> On Dienstag, 1. September 2020 14:25:31 CEST Richard Katsch wrote:
>>> Many thanks Matthias,Hugh.
>>>
>>> The case doesn’t seem to have any direct method of fastening to a heatsink
>>> but something can be arranged I’m sure. The manufacturer’s documents don’t
>>> suggest anything. They do say that the base temperature can be up to 85
>>> degrees C which seems awfully hot.
>>>
>>> I haven’t measured the temperature other than a finger test that indicates
>>> that more than 5 secs is uncomfortably hot. At that point I switched it off.
>>
>> I wouldn't say "uncomfortably hot" is a cause for alarm. The lamp housing in
>> the LPRO-101 runs at 100°C, the resonance chamber at roughly the same temp.
>> You need Rb vapor at around that temperature (if that Rb is a traditional
>> construction) for the system to work.
>>
>> Anyway, what else can you do? If it's broken, the chances of repairing it are
>> slim. If it works, no way to find out if it's in spec unless you run it. Just
>> watch for smoke ;)
>>
>> BR,
>> Matthias
>>
>>>
>>> Best regards
>>> Richard
>>>
>>> On Tue, 1 Sep 2020 at 19:02, Hugh Blemings <hugh@blemings.org> wrote:
>>>> Hiya,
>>>>
>>>>
>>>> I'd wondered this also - my (limited) experience of Rbs are the
>>>>
>>>> LPROs and they definitely need a heatsink (I have a nice story to
>>>>
>>>> share about getting some made one day which I must write up, but I
>>>>
>>>> digress)
>>>>
>>>>
>>>> Curious, I dug around a bit and only data I can find is a two
>>>>
>>>> pager on the manufacturer website -
>>>>
>>>> https://www.quartzlock.com/product/Rubidium/rubidium-oscillators/E10-MRX
>>>>
>>>>
>>>> The packaging itself doesn't appear to be designed with
>>>>
>>>> heatsinking in mind, but perhaps there is an implicit expectation
>>>>
>>>> of convection cooling at least to keep it manageable ?
>>>>
>>>>
>>>> Any idea what case temperature you're seeing @Richard ?
>>>>
>>>>
>>>>
>>>>
>>>> Cheers,
>>>>
>>>>
>>>> Hugh
>>>>
>>>>
>>>>
>>>>
>>>>
>>>>
>>>>
>>>> On 1/9/20 5:01 pm, Matthias Welwarsky
>>>>
>>>> wrote:
>>>>
>>>>
>>>>
>>>>
>>>>
>>>>
>>>> On Dienstag, 1. September 2020 07:15:34 CEST Richard Katsch wrote:
>>>>
>>>>
>>>>
>>>>
>>>>
>>>> Hello All,
>>>>
>>>> I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to
>>>> lock
>>>>
>>>> and produce a nice sine wave that stays in phase with my Trimble Tbolt for
>>>>
>>>> a time which exceeds my attention span!!!
>>>>
>>>> It does however get hot in operation. As this is my first experience with
>>>> a
>>>>
>>>> Rb standard I don’t know whether this is normal or indicates that I need a
>>>>
>>>> heat sink.
>>>>
>>>> Any comments would be appreciated.
>>>>
>>>>
>>>>
>>>>
>>>>
>>>>
>>>> They're supposed to run hot, but excess temperature will degrade the
>>>> stability
>>>>
>>>> and shorten the lifetime of the electronics. It's hard to recommend what
>>>>
>>>> heatsink they require if you have no documentation. Typically you'll need
>>>> one.
>>>>
>>>> My LPRO-101 need a heatsink with < 2K/W of thermal resistance to keep the
>>>>
>>>> temperature in check.
>>>>
>>>>
>>>>
>>>> BR,
>>>>
>>>> Matthias
>>>>
>>>>
>>>>
>>>>
>>>>
>>>>
>>>>
>>>> Regards
>>>>
>>>> Richard Katsch
>>>>
>>>> VK2EIK
>>>>
>>>> _______________________________________________
>>>>
>>>> time-nuts mailing list -- time-nuts@lists.febo.com
>>>>
>>>> To unsubscribe, go to
>>>>
>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow
>>>>
>>>> the instructions there.
>>>>
>>>>
>>>>
>>>>
>>>>
>>>>
>>>>
>>>>
>>>>
>>>>
>>>>
>>>>
>>>>
>>>>
>>>> _______________________________________________
>>>>
>>>> time-nuts mailing list -- time-nuts@lists.febo.com
>>>>
>>>> To unsubscribe, go to
>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
>>>>
>>>> and follow the instructions there.
>>
>>
>>
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@lists.febo.com
>> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
>> and follow the instructions there.
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.
DW
Dana Whitlow
Tue, Sep 1, 2020 11:04 PM
Taka,
There's not much really wrong in making the oven(s) work harder to maintain
the desired Rb cell
temperatures because of a cooler baseplate temperature. If the thermal
design is done well, nothing
will run appreciably hotter and most of the electronics will run cooler.
However, it is my strong belief that forced cooling with a fan will make
the baseplate temperature
fluctuate more as well as faster in response to local temperature
variations, which seems likely to
lead to poorer frequency stability on the time scale of minutes. This
effect could be mitigated by
adding thermal mass in the form of a thick aluminum block between the
baseplate and the heat
sink fins. However, doing so would make the warmup time (from a cold
start) of the Rb longer,
which may or may not bother you.
Dana
On Tue, Sep 1, 2020 at 4:16 PM Taka Kamiya via time-nuts <
time-nuts@lists.febo.com> wrote:
Will you elaborate on fan? Just about everything I've read, use of fan
was discouraged. Natural convection over heat sink fins were recommended.
I've done limited testing. When fan cooling, indeed temperature will go
down, but current consumption stays higher indicating heating is often
conducted to keep it hot inside. How do I reconciliate these opposing
conditions in actual use?
(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG
On Tuesday, September 1, 2020, 2:08:04 PM EDT, Bob kb8tq <
kb8tq@n1k.org> wrote:
Hi
In this case there may not be much that can be done. In the more general
case
( = a unit with a surface that mates to a heatsink) , it’s well worth
doing something.
I have an unfortunate lot of empirical evidence (on FRK’s and LPRO’s )
showing that
a “no heatsink” setup is one with a short life. With no heatsink, a couple
of years is
doing well on a surplus unit. With a heatsink, > 10 years is quite
possible.
How much heatsink? Is the next question. If you can get the baseplate down
into the
40’s (C) that’s doing pretty well. A fan plus heatsink may be needed to
get this done.
If a fan is involved, mag field could become an issue ( = Rb’s are
sensitive to mag
field, as the fan speed changes the field may change).
=====
Indeed this is not always the official word from the Rb manufacturers.
I’ve spent
considerable effort drinking beer with many of their marketing and
engineering
guys. The conversation is not quite the same when marketing it not at
the table.
I believe that they simply don’t have a body of long term data “in house”
to study.
The fact that I’m not the only one seeing Rb’s die early is backed up on
eBay.
From time to time, you can find pretty large batches of “for parts” Rb’s.
Date codes
often are not very far back in time.
If they get cheap enough (like ~ $30) they can be fun to play with. You
can pretty
quickly work out that the electrolytic caps are the most likely failure
points. Cap
swap outs may only fix 2/3 of them, but it’s quick and easy. You now have
a pile
of $45 Rb’s ….
Bob
On Dienstag, 1. September 2020 14:25:31 CEST Richard Katsch wrote:
Many thanks Matthias,Hugh.
The case doesn’t seem to have any direct method of fastening to a
but something can be arranged I’m sure. The manufacturer’s documents
suggest anything. They do say that the base temperature can be up to 85
degrees C which seems awfully hot.
I haven’t measured the temperature other than a finger test that
that more than 5 secs is uncomfortably hot. At that point I switched it
I wouldn't say "uncomfortably hot" is a cause for alarm. The lamp
the LPRO-101 runs at 100°C, the resonance chamber at roughly the same
You need Rb vapor at around that temperature (if that Rb is a
construction) for the system to work.
Anyway, what else can you do? If it's broken, the chances of repairing
slim. If it works, no way to find out if it's in spec unless you run it.
watch for smoke ;)
BR,
Matthias
Best regards
Richard
On Tue, 1 Sep 2020 at 19:02, Hugh Blemings hugh@blemings.org wrote:
Hiya,
I'd wondered this also - my (limited) experience of Rbs are the
LPROs and they definitely need a heatsink (I have a nice story to
share about getting some made one day which I must write up, but I
digress)
Curious, I dug around a bit and only data I can find is a two
pager on the manufacturer website -
The packaging itself doesn't appear to be designed with
heatsinking in mind, but perhaps there is an implicit expectation
of convection cooling at least to keep it manageable ?
Any idea what case temperature you're seeing @Richard ?
Cheers,
Hugh
On 1/9/20 5:01 pm, Matthias Welwarsky
wrote:
On Dienstag, 1. September 2020 07:15:34 CEST Richard Katsch wrote:
Hello All,
I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to
lock
and produce a nice sine wave that stays in phase with my Trimble Tbolt
a time which exceeds my attention span!!!
It does however get hot in operation. As this is my first experience
a
Rb standard I don’t know whether this is normal or indicates that I
heat sink.
Any comments would be appreciated.
They're supposed to run hot, but excess temperature will degrade the
stability
and shorten the lifetime of the electronics. It's hard to recommend
heatsink they require if you have no documentation. Typically you'll
one.
My LPRO-101 need a heatsink with < 2K/W of thermal resistance to keep
and follow the instructions there.
Taka,
There's not much really wrong in making the oven(s) work harder to maintain
the desired Rb cell
temperatures because of a cooler baseplate temperature. If the thermal
design is done well, nothing
will run appreciably hotter and most of the electronics will run cooler.
However, it is my strong belief that forced cooling with a fan will make
the baseplate temperature
fluctuate more as well as faster in response to local temperature
variations, which seems likely to
lead to poorer frequency stability on the time scale of minutes. This
effect could be mitigated by
adding thermal mass in the form of a thick aluminum block between the
baseplate and the heat
sink fins. However, doing so would make the warmup time (from a cold
start) of the Rb longer,
which may or may not bother you.
Dana
On Tue, Sep 1, 2020 at 4:16 PM Taka Kamiya via time-nuts <
time-nuts@lists.febo.com> wrote:
> Will you elaborate on fan? Just about everything I've read, use of fan
> was discouraged. Natural convection over heat sink fins were recommended.
> I've done limited testing. When fan cooling, indeed temperature will go
> down, but current consumption stays higher indicating heating is often
> conducted to keep it hot inside. How do I reconciliate these opposing
> conditions in actual use?
>
> ---------------------------------------
> (Mr.) Taka Kamiya
> KB4EMF / ex JF2DKG
>
>
> On Tuesday, September 1, 2020, 2:08:04 PM EDT, Bob kb8tq <
> kb8tq@n1k.org> wrote:
>
> Hi
>
> In this case there may not be much that can be done. In the more general
> case
> ( = a unit with a surface that mates to a heatsink) , it’s well worth
> doing something.
>
> I have an unfortunate lot of empirical evidence (on FRK’s and LPRO’s )
> showing that
> a “no heatsink” setup is one with a short life. With no heatsink, a couple
> of years is
> doing well on a surplus unit. With a heatsink, > 10 years is quite
> possible.
>
> How much heatsink? Is the next question. If you can get the baseplate down
> into the
> 40’s (C) that’s doing pretty well. A fan plus heatsink may be needed to
> get this done.
> If a fan is involved, mag field could become an issue ( = Rb’s are
> sensitive to mag
> field, as the fan speed changes the field *may* change).
>
> =====
>
> Indeed this is not always the official word from the Rb manufacturers.
> I’ve spent
> considerable effort drinking beer with many of their marketing and
> engineering
> guys. The conversation is not *quite* the same when marketing it not at
> the table.
> I believe that they simply don’t have a body of long term data “in house”
> to study.
>
> The fact that I’m not the only one seeing Rb’s die early is backed up on
> eBay.
> From time to time, you can find pretty large batches of “for parts” Rb’s.
> Date codes
> often are *not* very far back in time.
>
> If they get cheap enough (like ~ $30) they can be fun to play with. You
> can pretty
> quickly work out that the electrolytic caps are the most likely failure
> points. Cap
> swap outs may only fix 2/3 of them, but it’s quick and easy. You now have
> a pile
> of $45 Rb’s ….
>
> Bob
>
>
>
>
> > On Sep 1, 2020, at 9:23 AM, Matthias Welwarsky <time-nuts@welwarsky.de>
> wrote:
> >
> > On Dienstag, 1. September 2020 14:25:31 CEST Richard Katsch wrote:
> >> Many thanks Matthias,Hugh.
> >>
> >> The case doesn’t seem to have any direct method of fastening to a
> heatsink
> >> but something can be arranged I’m sure. The manufacturer’s documents
> don’t
> >> suggest anything. They do say that the base temperature can be up to 85
> >> degrees C which seems awfully hot.
> >>
> >> I haven’t measured the temperature other than a finger test that
> indicates
> >> that more than 5 secs is uncomfortably hot. At that point I switched it
> off.
> >
> > I wouldn't say "uncomfortably hot" is a cause for alarm. The lamp
> housing in
> > the LPRO-101 runs at 100°C, the resonance chamber at roughly the same
> temp.
> > You need Rb vapor at around that temperature (if that Rb is a
> traditional
> > construction) for the system to work.
> >
> > Anyway, what else can you do? If it's broken, the chances of repairing
> it are
> > slim. If it works, no way to find out if it's in spec unless you run it.
> Just
> > watch for smoke ;)
> >
> > BR,
> > Matthias
> >
> >>
> >> Best regards
> >> Richard
> >>
> >> On Tue, 1 Sep 2020 at 19:02, Hugh Blemings <hugh@blemings.org> wrote:
> >>> Hiya,
> >>>
> >>>
> >>> I'd wondered this also - my (limited) experience of Rbs are the
> >>>
> >>> LPROs and they definitely need a heatsink (I have a nice story to
> >>>
> >>> share about getting some made one day which I must write up, but I
> >>>
> >>> digress)
> >>>
> >>>
> >>> Curious, I dug around a bit and only data I can find is a two
> >>>
> >>> pager on the manufacturer website -
> >>>
> >>>
> https://www.quartzlock.com/product/Rubidium/rubidium-oscillators/E10-MRX
> >>>
> >>>
> >>> The packaging itself doesn't appear to be designed with
> >>>
> >>> heatsinking in mind, but perhaps there is an implicit expectation
> >>>
> >>> of convection cooling at least to keep it manageable ?
> >>>
> >>>
> >>> Any idea what case temperature you're seeing @Richard ?
> >>>
> >>>
> >>>
> >>>
> >>> Cheers,
> >>>
> >>>
> >>> Hugh
> >>>
> >>>
> >>>
> >>>
> >>>
> >>>
> >>>
> >>> On 1/9/20 5:01 pm, Matthias Welwarsky
> >>>
> >>> wrote:
> >>>
> >>>
> >>>
> >>>
> >>>
> >>>
> >>> On Dienstag, 1. September 2020 07:15:34 CEST Richard Katsch wrote:
> >>>
> >>>
> >>>
> >>>
> >>>
> >>> Hello All,
> >>>
> >>> I have acquired a Quartzlock E10-MRX Rb 10 MHz standard. It appears to
> >>> lock
> >>>
> >>> and produce a nice sine wave that stays in phase with my Trimble Tbolt
> for
> >>>
> >>> a time which exceeds my attention span!!!
> >>>
> >>> It does however get hot in operation. As this is my first experience
> with
> >>> a
> >>>
> >>> Rb standard I don’t know whether this is normal or indicates that I
> need a
> >>>
> >>> heat sink.
> >>>
> >>> Any comments would be appreciated.
> >>>
> >>>
> >>>
> >>>
> >>>
> >>>
> >>> They're supposed to run hot, but excess temperature will degrade the
> >>> stability
> >>>
> >>> and shorten the lifetime of the electronics. It's hard to recommend
> what
> >>>
> >>> heatsink they require if you have no documentation. Typically you'll
> need
> >>> one.
> >>>
> >>> My LPRO-101 need a heatsink with < 2K/W of thermal resistance to keep
> the
> >>>
> >>> temperature in check.
> >>>
> >>>
> >>>
> >>> BR,
> >>>
> >>> Matthias
> >>>
> >>>
> >>>
> >>>
> >>>
> >>>
> >>>
> >>> Regards
> >>>
> >>> Richard Katsch
> >>>
> >>> VK2EIK
> >>>
> >>> _______________________________________________
> >>>
> >>> time-nuts mailing list -- time-nuts@lists.febo.com
> >>>
> >>> To unsubscribe, go to
> >>>
> >>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and
> follow
> >>>
> >>> the instructions there.
> >>>
> >>>
> >>>
> >>>
> >>>
> >>>
> >>>
> >>>
> >>>
> >>>
> >>>
> >>>
> >>>
> >>>
> >>> _______________________________________________
> >>>
> >>> time-nuts mailing list -- time-nuts@lists.febo.com
> >>>
> >>> To unsubscribe, go to
> >>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> >>>
> >>> and follow the instructions there.
> >
> >
> >
> >
> >
> > _______________________________________________
> > time-nuts mailing list -- time-nuts@lists.febo.com
> > To unsubscribe, go to
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> > and follow the instructions there.
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.
>