PM6680 or 53131A for TimePod

SM
Scott McGrath
Sun, Nov 15, 2015 6:45 PM

Hi

Which counter would be best for use with time pod I have a chance to buy each at a reasonable prices i.e. < 1000 US

Thanks in advance Scott

Content by Scott
Typos by Siri

Hi Which counter would be best for use with time pod I have a chance to buy each at a reasonable prices i.e. < 1000 US Thanks in advance Scott Content by Scott Typos by Siri
BC
Bob Camp
Sun, Nov 15, 2015 9:59 PM

Hi

I guess the first question would be - If you have a Time Pod, what do you need a counter for?

Based on a guess that you need extended frequency range, either would do pretty well. My vote
would go to the HP for a couple of reasons:

  1. It’s more broadly supported in terms of driver software (which may or may not matter to you).

  2. It’s a more common item on the auction sites, so parts are easier to find.

  3. It shares some parts (like the power supply) with the (cheaper) 53181

No those are not great big reasons to go for one over the other. They are very similar devices. One
has an LCD that has a finite life. The other has a VFD that also has a finite life. Both are custom, that’s
a tie. Some like the command (button) arrangement on one over the other. I think it’s more a “what
are you used to” thing than anything else. On the 53131 you need to hit the Run button in some cases.

I have not watched the prices on the Fluke. I have kept track of the 53131’s. A “good price” on one
these days is below $600 ( = you may need to shop a bit, but they do show up).

Bob

On Nov 15, 2015, at 1:45 PM, Scott McGrath scmcgrath@gmail.com wrote:

Hi

Which counter would be best for use with time pod I have a chance to buy each at a reasonable prices i.e. < 1000 US

Thanks in advance Scott

Content by Scott
Typos by Siri


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi I guess the first question would be - If you have a Time Pod, what do you need a counter for? Based on a guess that you need extended frequency range, either would do pretty well. My vote would go to the HP for a couple of reasons: 1) It’s more broadly supported in terms of driver software (which may or may not matter to you). 2) It’s a more common item on the auction sites, so parts are easier to find. 3) It shares some parts (like the power supply) with the (cheaper) 53181 No those are not great big reasons to go for one over the other. They are very similar devices. One has an LCD that has a finite life. The other has a VFD that also has a finite life. Both are custom, that’s a tie. Some like the command (button) arrangement on one over the other. I think it’s more a “what are you used to” thing than anything else. On the 53131 you need to hit the Run button in some cases. I have not watched the prices on the Fluke. I have kept track of the 53131’s. A “good price” on one these days is below $600 ( = you may need to shop a bit, but they do show up). Bob > On Nov 15, 2015, at 1:45 PM, Scott McGrath <scmcgrath@gmail.com> wrote: > > Hi > > Which counter would be best for use with time pod I have a chance to buy each at a reasonable prices i.e. < 1000 US > > Thanks in advance Scott > > Content by Scott > Typos by Siri > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
SM
Scott McGrath
Sun, Nov 15, 2015 11:10 PM

Thanks Bob

I'm using the actual time pod for paying activities.  This would be a counter for time nuttery using the time pod software.  In those cases Where I don't want to fire up the 5071.

I'm familiar with the Keysight counters and I have a lot of them into the high Ghz ranges

The 53131 is spec'd to 500ps and the  6680 is spec'd at 250ps

I've also been leaning in the 53131 direction but I'd like to have the best tool for the job at hand

Content by Scott
Typos by Siri

On Nov 15, 2015, at 1:59 PM, Bob Camp kb8tq@n1k.org wrote:

Hi

I guess the first question would be - If you have a Time Pod, what do you need a counter for?

Based on a guess that you need extended frequency range, either would do pretty well. My vote
would go to the HP for a couple of reasons:

  1. It’s more broadly supported in terms of driver software (which may or may not matter to you).

  2. It’s a more common item on the auction sites, so parts are easier to find.

  3. It shares some parts (like the power supply) with the (cheaper) 53181

No those are not great big reasons to go for one over the other. They are very similar devices. One
has an LCD that has a finite life. The other has a VFD that also has a finite life. Both are custom, that’s
a tie. Some like the command (button) arrangement on one over the other. I think it’s more a “what
are you used to” thing than anything else. On the 53131 you need to hit the Run button in some cases.

I have not watched the prices on the Fluke. I have kept track of the 53131’s. A “good price” on one
these days is below $600 ( = you may need to shop a bit, but they do show up).

Bob

On Nov 15, 2015, at 1:45 PM, Scott McGrath scmcgrath@gmail.com wrote:

Hi

Which counter would be best for use with time pod I have a chance to buy each at a reasonable prices i.e. < 1000 US

Thanks in advance Scott

Content by Scott
Typos by Siri


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Thanks Bob I'm using the actual time pod for paying activities. This would be a counter for time nuttery using the time pod software. In those cases Where I don't want to fire up the 5071. I'm familiar with the Keysight counters and I have a lot of them into the high Ghz ranges The 53131 is spec'd to 500ps and the 6680 is spec'd at 250ps I've also been leaning in the 53131 direction but I'd like to have the best tool for the job at hand Content by Scott Typos by Siri > On Nov 15, 2015, at 1:59 PM, Bob Camp <kb8tq@n1k.org> wrote: > > Hi > > I guess the first question would be - If you have a Time Pod, what do you need a counter for? > > Based on a guess that you need extended frequency range, either would do pretty well. My vote > would go to the HP for a couple of reasons: > > 1) It’s more broadly supported in terms of driver software (which may or may not matter to you). > > 2) It’s a more common item on the auction sites, so parts are easier to find. > > 3) It shares some parts (like the power supply) with the (cheaper) 53181 > > No those are not great big reasons to go for one over the other. They are very similar devices. One > has an LCD that has a finite life. The other has a VFD that also has a finite life. Both are custom, that’s > a tie. Some like the command (button) arrangement on one over the other. I think it’s more a “what > are you used to” thing than anything else. On the 53131 you need to hit the Run button in some cases. > > I have not watched the prices on the Fluke. I have kept track of the 53131’s. A “good price” on one > these days is below $600 ( = you may need to shop a bit, but they do show up). > > Bob > >> On Nov 15, 2015, at 1:45 PM, Scott McGrath <scmcgrath@gmail.com> wrote: >> >> Hi >> >> Which counter would be best for use with time pod I have a chance to buy each at a reasonable prices i.e. < 1000 US >> >> Thanks in advance Scott >> >> Content by Scott >> Typos by Siri >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
CC
Chris Caudle
Mon, Nov 16, 2015 12:08 AM

On Sun, November 15, 2015 3:59 pm, Bob Camp wrote:

I guess the first question would be - If you have a Time Pod, what do you
need a counter for?

Or was the original question a slip, and was supposed to be which counter
is good for use with TimeLab software?

--
Chris Caudle

On Sun, November 15, 2015 3:59 pm, Bob Camp wrote: > I guess the first question would be - If you have a Time Pod, what do you > need a counter for? Or was the original question a slip, and was supposed to be which counter is good for use with TimeLab software? -- Chris Caudle
SM
Scott McGrath
Mon, Nov 16, 2015 6:42 AM

More of a slip as I've noted I have a time pod for work but I'd like to add a small counter for time nuttery as power is not cheap in NE so running the 5371A for long runs can get expensive

Content by Scott
Typos by Siri

On Nov 15, 2015, at 4:08 PM, Chris Caudle chris@chriscaudle.org wrote:

On Sun, November 15, 2015 3:59 pm, Bob Camp wrote:
I guess the first question would be - If you have a Time Pod, what do you
need a counter for?

Or was the original question a slip, and was supposed to be which counter
is good for use with TimeLab software?

--
Chris Caudle


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

More of a slip as I've noted I have a time pod for work but I'd like to add a small counter for time nuttery as power is not cheap in NE so running the 5371A for long runs can get expensive Content by Scott Typos by Siri > On Nov 15, 2015, at 4:08 PM, Chris Caudle <chris@chriscaudle.org> wrote: > >> On Sun, November 15, 2015 3:59 pm, Bob Camp wrote: >> I guess the first question would be - If you have a Time Pod, what do you >> need a counter for? > > Or was the original question a slip, and was supposed to be which counter > is good for use with TimeLab software? > > -- > Chris Caudle > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
LM
Larry McDavid
Mon, Nov 16, 2015 7:56 AM

Why would you want to use a 150 ps 53131 instead of a counter/timer like
the Agilent/Keysight 53230A that provides 20 ps resolution? The TimePod
my have better noise specs but it is a really dedicated, specific-use
device.

Larry

On 11/15/2015 10:42 PM, Scott McGrath wrote:

More of a slip as I've noted I have a time pod for work but I'd like to add a small counter for time nuttery as power is not cheap in NE so running the 5371A for long runs can get expensive

Content by Scott
Typos by Siri

On Nov 15, 2015, at 4:08 PM, Chris Caudle chris@chriscaudle.org wrote:

On Sun, November 15, 2015 3:59 pm, Bob Camp wrote:
I guess the first question would be - If you have a Time Pod, what do you
need a counter for?

Or was the original question a slip, and was supposed to be which counter
is good for use with TimeLab software?

--
Chris Caudle


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
Best wishes,

Larry McDavid W6FUB
Anaheim, California  (SE of Los Angeles, near Disneyland)

Why would you want to use a 150 ps 53131 instead of a counter/timer like the Agilent/Keysight 53230A that provides 20 ps resolution? The TimePod my have better noise specs but it is a really dedicated, specific-use device. Larry On 11/15/2015 10:42 PM, Scott McGrath wrote: > More of a slip as I've noted I have a time pod for work but I'd like to add a small counter for time nuttery as power is not cheap in NE so running the 5371A for long runs can get expensive > > Content by Scott > Typos by Siri > >> On Nov 15, 2015, at 4:08 PM, Chris Caudle <chris@chriscaudle.org> wrote: >> >>> On Sun, November 15, 2015 3:59 pm, Bob Camp wrote: >>> I guess the first question would be - If you have a Time Pod, what do you >>> need a counter for? >> >> Or was the original question a slip, and was supposed to be which counter >> is good for use with TimeLab software? >> >> -- >> Chris Caudle >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > -- Best wishes, Larry McDavid W6FUB Anaheim, California (SE of Los Angeles, near Disneyland)
CC
Chris Caudle
Mon, Nov 16, 2015 2:20 PM

On Mon, November 16, 2015 12:42 am, Scott McGrath wrote:

More of a slip as I've noted I have a time pod for work but I'd like to
add a small counter for time nuttery as power is not cheap in NE so
running the 5371A for long runs can get expensive

I don't follow.  Why would you not use the TimePod with a lower power
reference?  It isn't like a TimePod only works with a cesium clock as the
reference source.

--
Chris Caudle

On Mon, November 16, 2015 12:42 am, Scott McGrath wrote: > More of a slip as I've noted I have a time pod for work but I'd like to > add a small counter for time nuttery as power is not cheap in NE so > running the 5371A for long runs can get expensive I don't follow. Why would you not use the TimePod with a lower power reference? It isn't like a TimePod only works with a cesium clock as the reference source. -- Chris Caudle
A
Angus
Mon, Nov 16, 2015 9:32 PM

Hi,

PM6681's turn up at quite a good price too, and you get a lot of
extras over the 6680 - does anyone know if Timelab will work with a
6681?

Angus.

On Sun, 15 Nov 2015 15:10:42 -0800, you wrote:

Thanks Bob

I'm using the actual time pod for paying activities.  This would be a counter for time nuttery using the time pod software.  In those cases Where I don't want to fire up the 5071.

I'm familiar with the Keysight counters and I have a lot of them into the high Ghz ranges

The 53131 is spec'd to 500ps and the  6680 is spec'd at 250ps

I've also been leaning in the 53131 direction but I'd like to have the best tool for the job at hand

Content by Scott
Typos by Siri

On Nov 15, 2015, at 1:59 PM, Bob Camp kb8tq@n1k.org wrote:

Hi

I guess the first question would be - If you have a Time Pod, what do you need a counter for?

Based on a guess that you need extended frequency range, either would do pretty well. My vote
would go to the HP for a couple of reasons:

  1. It’s more broadly supported in terms of driver software (which may or may not matter to you).

  2. It’s a more common item on the auction sites, so parts are easier to find.

  3. It shares some parts (like the power supply) with the (cheaper) 53181

No those are not great big reasons to go for one over the other. They are very similar devices. One
has an LCD that has a finite life. The other has a VFD that also has a finite life. Both are custom, that’s
a tie. Some like the command (button) arrangement on one over the other. I think it’s more a “what
are you used to” thing than anything else. On the 53131 you need to hit the Run button in some cases.

I have not watched the prices on the Fluke. I have kept track of the 53131’s. A “good price” on one
these days is below $600 ( = you may need to shop a bit, but they do show up).

Bob

On Nov 15, 2015, at 1:45 PM, Scott McGrath scmcgrath@gmail.com wrote:

Hi

Which counter would be best for use with time pod I have a chance to buy each at a reasonable prices i.e. < 1000 US

Thanks in advance Scott

Content by Scott
Typos by Siri


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi, PM6681's turn up at quite a good price too, and you get a lot of extras over the 6680 - does anyone know if Timelab will work with a 6681? Angus. On Sun, 15 Nov 2015 15:10:42 -0800, you wrote: >Thanks Bob > >I'm using the actual time pod for paying activities. This would be a counter for time nuttery using the time pod software. In those cases Where I don't want to fire up the 5071. > >I'm familiar with the Keysight counters and I have a lot of them into the high Ghz ranges > >The 53131 is spec'd to 500ps and the 6680 is spec'd at 250ps > >I've also been leaning in the 53131 direction but I'd like to have the best tool for the job at hand > >Content by Scott >Typos by Siri > >> On Nov 15, 2015, at 1:59 PM, Bob Camp <kb8tq@n1k.org> wrote: >> >> Hi >> >> I guess the first question would be - If you have a Time Pod, what do you need a counter for? >> >> Based on a guess that you need extended frequency range, either would do pretty well. My vote >> would go to the HP for a couple of reasons: >> >> 1) It’s more broadly supported in terms of driver software (which may or may not matter to you). >> >> 2) It’s a more common item on the auction sites, so parts are easier to find. >> >> 3) It shares some parts (like the power supply) with the (cheaper) 53181 >> >> No those are not great big reasons to go for one over the other. They are very similar devices. One >> has an LCD that has a finite life. The other has a VFD that also has a finite life. Both are custom, that’s >> a tie. Some like the command (button) arrangement on one over the other. I think it’s more a “what >> are you used to” thing than anything else. On the 53131 you need to hit the Run button in some cases. >> >> I have not watched the prices on the Fluke. I have kept track of the 53131’s. A “good price” on one >> these days is below $600 ( = you may need to shop a bit, but they do show up). >> >> Bob >> >>> On Nov 15, 2015, at 1:45 PM, Scott McGrath <scmcgrath@gmail.com> wrote: >>> >>> Hi >>> >>> Which counter would be best for use with time pod I have a chance to buy each at a reasonable prices i.e. < 1000 US >>> >>> Thanks in advance Scott >>> >>> Content by Scott >>> Typos by Siri >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >_______________________________________________ >time-nuts mailing list -- time-nuts@febo.com >To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >and follow the instructions there.
BC
Bob Camp
Tue, Nov 17, 2015 1:27 AM

Hi

At leas on the spec sheet I found, both counters have a single shot of 500 ps. The
6680 “improves” as the averaging goes up.

http://isurplus.com.au/specifications/Philips%20PM6680%20Specifications.pdf

Section 13-8 has a lot of “500 ps / sqrt(N)" in it. If N=1 (single shot) you get 500 ps. HP stops the
process when resolution and random bounce are roughly equal. Fluke gives you another bit of
noise to possibly average out under some conditions.

At least the way I read the data sheets, the two counters are very similar.

Bob

On Nov 15, 2015, at 6:10 PM, Scott McGrath scmcgrath@gmail.com wrote:

Thanks Bob

I'm using the actual time pod for paying activities.  This would be a counter for time nuttery using the time pod software.  In those cases Where I don't want to fire up the 5071.

I'm familiar with the Keysight counters and I have a lot of them into the high Ghz ranges

The 53131 is spec'd to 500ps and the  6680 is spec'd at 250ps

I've also been leaning in the 53131 direction but I'd like to have the best tool for the job at hand

Content by Scott
Typos by Siri

On Nov 15, 2015, at 1:59 PM, Bob Camp kb8tq@n1k.org wrote:

Hi

I guess the first question would be - If you have a Time Pod, what do you need a counter for?

Based on a guess that you need extended frequency range, either would do pretty well. My vote
would go to the HP for a couple of reasons:

  1. It’s more broadly supported in terms of driver software (which may or may not matter to you).

  2. It’s a more common item on the auction sites, so parts are easier to find.

  3. It shares some parts (like the power supply) with the (cheaper) 53181

No those are not great big reasons to go for one over the other. They are very similar devices. One
has an LCD that has a finite life. The other has a VFD that also has a finite life. Both are custom, that’s
a tie. Some like the command (button) arrangement on one over the other. I think it’s more a “what
are you used to” thing than anything else. On the 53131 you need to hit the Run button in some cases.

I have not watched the prices on the Fluke. I have kept track of the 53131’s. A “good price” on one
these days is below $600 ( = you may need to shop a bit, but they do show up).

Bob

On Nov 15, 2015, at 1:45 PM, Scott McGrath scmcgrath@gmail.com wrote:

Hi

Which counter would be best for use with time pod I have a chance to buy each at a reasonable prices i.e. < 1000 US

Thanks in advance Scott

Content by Scott
Typos by Siri


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi At leas on the spec sheet I found, both counters have a single shot of 500 ps. The 6680 “improves” as the averaging goes up. http://isurplus.com.au/specifications/Philips%20PM6680%20Specifications.pdf Section 13-8 has a lot of “500 ps / sqrt(N)" in it. If N=1 (single shot) you get 500 ps. HP stops the process when resolution and random bounce are roughly equal. Fluke gives you another bit of noise to possibly average out under some conditions. At least the way I read the data sheets, the two counters are very similar. Bob > On Nov 15, 2015, at 6:10 PM, Scott McGrath <scmcgrath@gmail.com> wrote: > > Thanks Bob > > I'm using the actual time pod for paying activities. This would be a counter for time nuttery using the time pod software. In those cases Where I don't want to fire up the 5071. > > I'm familiar with the Keysight counters and I have a lot of them into the high Ghz ranges > > The 53131 is spec'd to 500ps and the 6680 is spec'd at 250ps > > I've also been leaning in the 53131 direction but I'd like to have the best tool for the job at hand > > Content by Scott > Typos by Siri > >> On Nov 15, 2015, at 1:59 PM, Bob Camp <kb8tq@n1k.org> wrote: >> >> Hi >> >> I guess the first question would be - If you have a Time Pod, what do you need a counter for? >> >> Based on a guess that you need extended frequency range, either would do pretty well. My vote >> would go to the HP for a couple of reasons: >> >> 1) It’s more broadly supported in terms of driver software (which may or may not matter to you). >> >> 2) It’s a more common item on the auction sites, so parts are easier to find. >> >> 3) It shares some parts (like the power supply) with the (cheaper) 53181 >> >> No those are not great big reasons to go for one over the other. They are very similar devices. One >> has an LCD that has a finite life. The other has a VFD that also has a finite life. Both are custom, that’s >> a tie. Some like the command (button) arrangement on one over the other. I think it’s more a “what >> are you used to” thing than anything else. On the 53131 you need to hit the Run button in some cases. >> >> I have not watched the prices on the Fluke. I have kept track of the 53131’s. A “good price” on one >> these days is below $600 ( = you may need to shop a bit, but they do show up). >> >> Bob >> >>> On Nov 15, 2015, at 1:45 PM, Scott McGrath <scmcgrath@gmail.com> wrote: >>> >>> Hi >>> >>> Which counter would be best for use with time pod I have a chance to buy each at a reasonable prices i.e. < 1000 US >>> >>> Thanks in advance Scott >>> >>> Content by Scott >>> Typos by Siri >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
SM
Scott McGrath
Tue, Nov 17, 2015 1:45 AM

5371a is a TI analyzer  The time pod itself is used for paying work so it's not available for time nuttery.

Content by Scott
Typos by Siri

On Nov 16, 2015, at 6:20 AM, Chris Caudle chris@chriscaudle.org wrote:

On Mon, November 16, 2015 12:42 am, Scott McGrath wrote:
More of a slip as I've noted I have a time pod for work but I'd like to
add a small counter for time nuttery as power is not cheap in NE so
running the 5371A for long runs can get expensive

I don't follow.  Why would you not use the TimePod with a lower power
reference?  It isn't like a TimePod only works with a cesium clock as the
reference source.

--
Chris Caudle


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

5371a is a TI analyzer The time pod itself is used for paying work so it's not available for time nuttery. Content by Scott Typos by Siri > On Nov 16, 2015, at 6:20 AM, Chris Caudle <chris@chriscaudle.org> wrote: > >> On Mon, November 16, 2015 12:42 am, Scott McGrath wrote: >> More of a slip as I've noted I have a time pod for work but I'd like to >> add a small counter for time nuttery as power is not cheap in NE so >> running the 5371A for long runs can get expensive > > I don't follow. Why would you not use the TimePod with a lower power > reference? It isn't like a TimePod only works with a cesium clock as the > reference source. > > -- > Chris Caudle > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
MD
Magnus Danielson
Tue, Nov 17, 2015 6:51 AM

Bob,

The single-shot does not improve, it's 500 ps, but the averaging of N
samples give you that result. The 53131 does similar tricks.

Cheers,
Magnus

On 11/17/2015 02:27 AM, Bob Camp wrote:

Hi

At leas on the spec sheet I found, both counters have a single shot of 500 ps. The
6680 “improves” as the averaging goes up.

http://isurplus.com.au/specifications/Philips%20PM6680%20Specifications.pdf

Section 13-8 has a lot of “500 ps / sqrt(N)" in it. If N=1 (single shot) you get 500 ps. HP stops the
process when resolution and random bounce are roughly equal. Fluke gives you another bit of
noise to possibly average out under some conditions.

At least the way I read the data sheets, the two counters are very similar.

Bob

On Nov 15, 2015, at 6:10 PM, Scott McGrath scmcgrath@gmail.com wrote:

Thanks Bob

I'm using the actual time pod for paying activities.  This would be a counter for time nuttery using the time pod software.  In those cases Where I don't want to fire up the 5071.

I'm familiar with the Keysight counters and I have a lot of them into the high Ghz ranges

The 53131 is spec'd to 500ps and the  6680 is spec'd at 250ps

I've also been leaning in the 53131 direction but I'd like to have the best tool for the job at hand

Content by Scott
Typos by Siri

On Nov 15, 2015, at 1:59 PM, Bob Camp kb8tq@n1k.org wrote:

Hi

I guess the first question would be - If you have a Time Pod, what do you need a counter for?

Based on a guess that you need extended frequency range, either would do pretty well. My vote
would go to the HP for a couple of reasons:

  1. It’s more broadly supported in terms of driver software (which may or may not matter to you).

  2. It’s a more common item on the auction sites, so parts are easier to find.

  3. It shares some parts (like the power supply) with the (cheaper) 53181

No those are not great big reasons to go for one over the other. They are very similar devices. One
has an LCD that has a finite life. The other has a VFD that also has a finite life. Both are custom, that’s
a tie. Some like the command (button) arrangement on one over the other. I think it’s more a “what
are you used to” thing than anything else. On the 53131 you need to hit the Run button in some cases.

I have not watched the prices on the Fluke. I have kept track of the 53131’s. A “good price” on one
these days is below $600 ( = you may need to shop a bit, but they do show up).

Bob

On Nov 15, 2015, at 1:45 PM, Scott McGrath scmcgrath@gmail.com wrote:

Hi

Which counter would be best for use with time pod I have a chance to buy each at a reasonable prices i.e. < 1000 US

Thanks in advance Scott

Content by Scott
Typos by Siri


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Bob, The single-shot does not improve, it's 500 ps, but the averaging of N samples give you that result. The 53131 does similar tricks. Cheers, Magnus On 11/17/2015 02:27 AM, Bob Camp wrote: > Hi > > At leas on the spec sheet I found, both counters have a single shot of 500 ps. The > 6680 “improves” as the averaging goes up. > > http://isurplus.com.au/specifications/Philips%20PM6680%20Specifications.pdf > > Section 13-8 has a lot of “500 ps / sqrt(N)" in it. If N=1 (single shot) you get 500 ps. HP stops the > process when resolution and random bounce are roughly equal. Fluke gives you another bit of > noise to possibly average out under some conditions. > > At least the way I read the data sheets, the two counters are very similar. > > Bob > >> On Nov 15, 2015, at 6:10 PM, Scott McGrath <scmcgrath@gmail.com> wrote: >> >> Thanks Bob >> >> I'm using the actual time pod for paying activities. This would be a counter for time nuttery using the time pod software. In those cases Where I don't want to fire up the 5071. >> >> I'm familiar with the Keysight counters and I have a lot of them into the high Ghz ranges >> >> The 53131 is spec'd to 500ps and the 6680 is spec'd at 250ps >> >> I've also been leaning in the 53131 direction but I'd like to have the best tool for the job at hand >> >> Content by Scott >> Typos by Siri >> >>> On Nov 15, 2015, at 1:59 PM, Bob Camp <kb8tq@n1k.org> wrote: >>> >>> Hi >>> >>> I guess the first question would be - If you have a Time Pod, what do you need a counter for? >>> >>> Based on a guess that you need extended frequency range, either would do pretty well. My vote >>> would go to the HP for a couple of reasons: >>> >>> 1) It’s more broadly supported in terms of driver software (which may or may not matter to you). >>> >>> 2) It’s a more common item on the auction sites, so parts are easier to find. >>> >>> 3) It shares some parts (like the power supply) with the (cheaper) 53181 >>> >>> No those are not great big reasons to go for one over the other. They are very similar devices. One >>> has an LCD that has a finite life. The other has a VFD that also has a finite life. Both are custom, that’s >>> a tie. Some like the command (button) arrangement on one over the other. I think it’s more a “what >>> are you used to” thing than anything else. On the 53131 you need to hit the Run button in some cases. >>> >>> I have not watched the prices on the Fluke. I have kept track of the 53131’s. A “good price” on one >>> these days is below $600 ( = you may need to shop a bit, but they do show up). >>> >>> Bob >>> >>>> On Nov 15, 2015, at 1:45 PM, Scott McGrath <scmcgrath@gmail.com> wrote: >>>> >>>> Hi >>>> >>>> Which counter would be best for use with time pod I have a chance to buy each at a reasonable prices i.e. < 1000 US >>>> >>>> Thanks in advance Scott >>>> >>>> Content by Scott >>>> Typos by Siri >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>> and follow the instructions there. >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
D
davidh
Tue, Nov 17, 2015 11:33 AM

Hi Angus,

I have three 6681s operating concurrently under a single instance of
timelab beta. It operates perfectly, even live n-cornered measurements.
Kudos to John!

The only downside (and I understand why it's this way) is that each
counter needs a unique GPIB interface, but this is a timelab requirement
and nothing to do with the counters.

My only fear regarding the counters is that I'll lose the cal one day,
so I'm keen to discover how we can calibrate them ourselves.

Cheers,

david

On 17/11/2015 7:32 AM, Angus wrote:

Hi,

PM6681's turn up at quite a good price too, and you get a lot of
extras over the 6680 - does anyone know if Timelab will work with a
6681?

Angus.

On Sun, 15 Nov 2015 15:10:42 -0800, you wrote:

Hi Angus, I have three 6681s operating concurrently under a single instance of timelab beta. It operates perfectly, even live n-cornered measurements. Kudos to John! The only downside (and I understand why it's this way) is that each counter needs a unique GPIB interface, but this is a timelab requirement and nothing to do with the counters. My only fear regarding the counters is that I'll lose the cal one day, so I'm keen to discover how we can calibrate them ourselves. Cheers, david On 17/11/2015 7:32 AM, Angus wrote: > Hi, > > PM6681's turn up at quite a good price too, and you get a lot of > extras over the 6680 - does anyone know if Timelab will work with a > 6681? > > Angus. > > > > On Sun, 15 Nov 2015 15:10:42 -0800, you wrote: >
AB
Azelio Boriani
Tue, Nov 17, 2015 7:34 PM

Not necessarily, if you follow the service manual's procedure, you can
replace the backup 3V cell safely, keeping the calibration data: the
problem is checking periodically the backup cell or replacing it
regularly (say, every 5 years).

On Tue, Nov 17, 2015 at 12:33 PM, davidh dhooke@gmail.com wrote:

Hi Angus,

I have three 6681s operating concurrently under a single instance of timelab
beta. It operates perfectly, even live n-cornered measurements. Kudos to
John!

The only downside (and I understand why it's this way) is that each counter
needs a unique GPIB interface, but this is a timelab requirement and nothing
to do with the counters.

My only fear regarding the counters is that I'll lose the cal one day, so
I'm keen to discover how we can calibrate them ourselves.

Cheers,

david

On 17/11/2015 7:32 AM, Angus wrote:

Hi,

PM6681's turn up at quite a good price too, and you get a lot of
extras over the 6680 - does anyone know if Timelab will work with a
6681?

Angus.

On Sun, 15 Nov 2015 15:10:42 -0800, you wrote:


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Not necessarily, if you follow the service manual's procedure, you can replace the backup 3V cell safely, keeping the calibration data: the problem is checking periodically the backup cell or replacing it regularly (say, every 5 years). On Tue, Nov 17, 2015 at 12:33 PM, davidh <dhooke@gmail.com> wrote: > > > Hi Angus, > > I have three 6681s operating concurrently under a single instance of timelab > beta. It operates perfectly, even live n-cornered measurements. Kudos to > John! > > The only downside (and I understand why it's this way) is that each counter > needs a unique GPIB interface, but this is a timelab requirement and nothing > to do with the counters. > > My only fear regarding the counters is that I'll lose the cal one day, so > I'm keen to discover how we can calibrate them ourselves. > > Cheers, > > david > > > On 17/11/2015 7:32 AM, Angus wrote: >> >> Hi, >> >> PM6681's turn up at quite a good price too, and you get a lot of >> extras over the 6680 - does anyone know if Timelab will work with a >> 6681? >> >> Angus. >> >> >> >> On Sun, 15 Nov 2015 15:10:42 -0800, you wrote: >> > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
AB
Azelio Boriani
Tue, Nov 17, 2015 8:30 PM

Not necessarily, if you follow the service manual's procedure, you can
replace the backup 3V cell safely, keeping the calibration data: the
problem is checking periodically the backup cell or replacing it
regularly (say, every 5 years).

On Tue, Nov 17, 2015 at 12:33 PM, davidh dhooke@gmail.com wrote:

Hi Angus,

I have three 6681s operating concurrently under a single instance of timelab
beta. It operates perfectly, even live n-cornered measurements. Kudos to
John!

The only downside (and I understand why it's this way) is that each counter
needs a unique GPIB interface, but this is a timelab requirement and nothing
to do with the counters.

My only fear regarding the counters is that I'll lose the cal one day, so
I'm keen to discover how we can calibrate them ourselves.

Cheers,

david

On 17/11/2015 7:32 AM, Angus wrote:

Hi,

PM6681's turn up at quite a good price too, and you get a lot of
extras over the 6680 - does anyone know if Timelab will work with a
6681?

Angus.

On Sun, 15 Nov 2015 15:10:42 -0800, you wrote:


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Not necessarily, if you follow the service manual's procedure, you can replace the backup 3V cell safely, keeping the calibration data: the problem is checking periodically the backup cell or replacing it regularly (say, every 5 years). On Tue, Nov 17, 2015 at 12:33 PM, davidh <dhooke@gmail.com> wrote: > > > Hi Angus, > > I have three 6681s operating concurrently under a single instance of timelab > beta. It operates perfectly, even live n-cornered measurements. Kudos to > John! > > The only downside (and I understand why it's this way) is that each counter > needs a unique GPIB interface, but this is a timelab requirement and nothing > to do with the counters. > > My only fear regarding the counters is that I'll lose the cal one day, so > I'm keen to discover how we can calibrate them ourselves. > > Cheers, > > david > > > On 17/11/2015 7:32 AM, Angus wrote: >> >> Hi, >> >> PM6681's turn up at quite a good price too, and you get a lot of >> extras over the 6680 - does anyone know if Timelab will work with a >> 6681? >> >> Angus. >> >> >> >> On Sun, 15 Nov 2015 15:10:42 -0800, you wrote: >> > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BC
Bob Camp
Wed, Nov 18, 2015 12:35 PM

Hi

The coin cell / backup battery swap out is something we probably will become more familiar
with on a lot of gear. The battery backed up RAM idea is now old enough that there is a large
population of test gear / radios / telecom gear out there with this “feature”. In some cases the
loss of the battery is a temporary issue. In a lot of others it’s a significant problem.

If you are buying a piece of gear that has important stuff in RAM, the big question is — has the
data been lost already? I have bought gear that had a good battery in it, but bad data. If the gear
comes up with “data lost” on the screen, that’s easy to spot. In most cases …not so easy at all.

Some gear might be configurable by normal means. Almost everything I’ve seen needs a “factory
only” shoot from a test set that probably no longer exists. Yes, there’s nothing magic in that test
set. The RAM just has bits in it. Figuring out what all the bits need to be without any documentation
is not easy.

One might hope that as the gear becomes obsolete, the information about what’s what would be
released to the public. Based on … errr …. on the job experience - not so likely. The data rarely
is documented in a “public compatible” fashion. One guy’s notes tell you what the test setup looks like.
Another set of notes go into the code. Both are buried in log books from who knows when. Beyond
this, someone actively has to agree to release “corporate IP”. The complex part of that is the fact
that the calibration techniques probably live on in a modern piece of gear. Not at all easy ….

Bob

Hi The coin cell / backup battery swap out is something we probably will become more familiar with on a *lot* of gear. The battery backed up RAM idea is now old enough that there is a large population of test gear / radios / telecom gear out there with this “feature”. In some cases the loss of the battery is a temporary issue. In a lot of others it’s a significant problem. If you are buying a piece of gear that has important stuff in RAM, the big question is — has the data been lost already? I have bought gear that had a good battery in it, but bad data. If the gear comes up with “data lost” on the screen, that’s easy to spot. In most cases …not so easy at all. Some gear might be configurable by normal means. Almost everything I’ve seen needs a “factory only” shoot from a test set that probably no longer exists. Yes, there’s nothing magic in that test set. The RAM just has bits in it. Figuring out what all the bits need to be without any documentation is not easy. One might hope that as the gear becomes obsolete, the information about what’s what would be released to the public. Based on … errr …. on the job experience - not so likely. The data rarely is documented in a “public compatible” fashion. One guy’s notes tell you what the test setup looks like. Another set of notes go into the code. Both are buried in log books from who knows when. Beyond this, someone actively has to agree to release “corporate IP”. The complex part of that is the fact that the calibration techniques probably live on in a modern piece of gear. Not at all easy …. Bob
MD
Magnus Danielson
Wed, Nov 18, 2015 10:18 PM

Bob,

To illustrate your point.

The original PM6681 calibration setup includes a PC, an ancient Philips
ISA-bus GPIB interface and an ancient Philips driver and a DOS program.
Collecting these and put them together to work will be an interesting
challenge.

A more modern variant of the calibration routines run can run on more
modern HW and LabView software. Philips PM5781 and Agilent 81112A pulse
generators is supported by those routines.

By now the production of the counter has stopped, as the counter core
ASIC ran out of stock.

The folks who designed and maintained them does not work there anymore.

Turns out the core setup requires a generator that creates a skewed
frequency such that you sweep over the interpolation phases. You tweak
the calibration value (3.86-4.50 ns in 0.02 ns steps) that you set over
GPIB using the *PUD command. You can check it yourself using *PUD?.
It sweeps over this range and checks for the value giving lowest RMS
value (SDEV result) and then sets this as the final value.

You can set the value using the command :SYST:UNPR;*PUD %s where %s is
the string, looking somewhat like this:

FACTORY CALIBRATED: %s%s, CALPLS 3.98 ns

The two %s in there is for some string and date, I just don't bother to
dig up a correct example.

Anyway, you can actually read-out the data and write it back without the
software. It should be possible to do something with the existing pieces
and it might be possible to actually get the calibration software
operational again.

I just don't have a CNT81 anymore to try this out on.

Cheers,
Magnus

On 11/18/2015 01:35 PM, Bob Camp wrote:

Hi

The coin cell / backup battery swap out is something we probably will become more familiar
with on a lot of gear. The battery backed up RAM idea is now old enough that there is a large
population of test gear / radios / telecom gear out there with this “feature”. In some cases the
loss of the battery is a temporary issue. In a lot of others it’s a significant problem.

If you are buying a piece of gear that has important stuff in RAM, the big question is — has the
data been lost already? I have bought gear that had a good battery in it, but bad data. If the gear
comes up with “data lost” on the screen, that’s easy to spot. In most cases …not so easy at all.

Some gear might be configurable by normal means. Almost everything I’ve seen needs a “factory
only” shoot from a test set that probably no longer exists. Yes, there’s nothing magic in that test
set. The RAM just has bits in it. Figuring out what all the bits need to be without any documentation
is not easy.

One might hope that as the gear becomes obsolete, the information about what’s what would be
released to the public. Based on … errr …. on the job experience - not so likely. The data rarely
is documented in a “public compatible” fashion. One guy’s notes tell you what the test setup looks like.
Another set of notes go into the code. Both are buried in log books from who knows when. Beyond
this, someone actively has to agree to release “corporate IP”. The complex part of that is the fact
that the calibration techniques probably live on in a modern piece of gear. Not at all easy ….

Bob


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Bob, To illustrate your point. The original PM6681 calibration setup includes a PC, an ancient Philips ISA-bus GPIB interface and an ancient Philips driver and a DOS program. Collecting these and put them together to work will be an interesting challenge. A more modern variant of the calibration routines run can run on more modern HW and LabView software. Philips PM5781 and Agilent 81112A pulse generators is supported by those routines. By now the production of the counter has stopped, as the counter core ASIC ran out of stock. The folks who designed and maintained them does not work there anymore. Turns out the core setup requires a generator that creates a skewed frequency such that you sweep over the interpolation phases. You tweak the calibration value (3.86-4.50 ns in 0.02 ns steps) that you set over GPIB using the *PUD command. You can check it yourself using *PUD?. It sweeps over this range and checks for the value giving lowest RMS value (SDEV result) and then sets this as the final value. You can set the value using the command :SYST:UNPR;*PUD %s where %s is the string, looking somewhat like this: FACTORY CALIBRATED: %s%s, CALPLS 3.98 ns The two %s in there is for some string and date, I just don't bother to dig up a correct example. Anyway, you can actually read-out the data and write it back without the software. It should be possible to do something with the existing pieces and it might be possible to actually get the calibration software operational again. I just don't have a CNT81 anymore to try this out on. Cheers, Magnus On 11/18/2015 01:35 PM, Bob Camp wrote: > Hi > > The coin cell / backup battery swap out is something we probably will become more familiar > with on a *lot* of gear. The battery backed up RAM idea is now old enough that there is a large > population of test gear / radios / telecom gear out there with this “feature”. In some cases the > loss of the battery is a temporary issue. In a lot of others it’s a significant problem. > > If you are buying a piece of gear that has important stuff in RAM, the big question is — has the > data been lost already? I have bought gear that had a good battery in it, but bad data. If the gear > comes up with “data lost” on the screen, that’s easy to spot. In most cases …not so easy at all. > > Some gear might be configurable by normal means. Almost everything I’ve seen needs a “factory > only” shoot from a test set that probably no longer exists. Yes, there’s nothing magic in that test > set. The RAM just has bits in it. Figuring out what all the bits need to be without any documentation > is not easy. > > One might hope that as the gear becomes obsolete, the information about what’s what would be > released to the public. Based on … errr …. on the job experience - not so likely. The data rarely > is documented in a “public compatible” fashion. One guy’s notes tell you what the test setup looks like. > Another set of notes go into the code. Both are buried in log books from who knows when. Beyond > this, someone actively has to agree to release “corporate IP”. The complex part of that is the fact > that the calibration techniques probably live on in a modern piece of gear. Not at all easy …. > > Bob > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >