Hi all,
Awhile back I posted about some mysterious 0.7 ns jumps in three different
masers (of the same brand) at three different locations around Australia.
Well we think we've found the problem. All three locations also have
in-room air conditioners of the same brand. These are used for cooling
only. When these units turn on, we think they induce a magnetic field from
the inrush current that briefly disrupts the maser. We don't think it's
electrical because moving to another phase did not change things.
These air conditioners are all quite close to the masers. Typically a metre
or 2 away.
Much was done to discover this, but the clincher was that when the weather
cooled enough at the southern most location (Hobart), we turned off the air
con (only heating was needed) and the problem vanished.
So there's a lesson here for all maser owners. The jump of 0.7 nsec is not
much, but it's huge for VLBI and for time-nuts.
Jim Palfreyman
Jim,
On 05/22/2016 03:58 AM, Jim Palfreyman wrote:
Hi all,
Awhile back I posted about some mysterious 0.7 ns jumps in three different
masers (of the same brand) at three different locations around Australia.
Well we think we've found the problem. All three locations also have
in-room air conditioners of the same brand. These are used for cooling
only. When these units turn on, we think they induce a magnetic field from
the inrush current that briefly disrupts the maser. We don't think it's
electrical because moving to another phase did not change things.
These air conditioners are all quite close to the masers. Typically a metre
or 2 away.
Much was done to discover this, but the clincher was that when the weather
cooled enough at the southern most location (Hobart), we turned off the air
con (only heating was needed) and the problem vanished.
So there's a lesson here for all maser owners. The jump of 0.7 nsec is not
much, but it's huge for VLBI and for time-nuts.
Good that you have found the offender, but have you been able to remedy
it by other means than turning the AC off?
I think others H-maser owners would love to know, and potentially the
vendor you have.
Cheers,
Magnus
Hi
That’s a pretty good example for the “why you don’t do your timescale based on
a single brand of gear / single setup” file.
Thanks for sharing!!
Bob
On May 21, 2016, at 9:58 PM, Jim Palfreyman jim77742@gmail.com wrote:
Hi all,
Awhile back I posted about some mysterious 0.7 ns jumps in three different
masers (of the same brand) at three different locations around Australia.
Well we think we've found the problem. All three locations also have
in-room air conditioners of the same brand. These are used for cooling
only. When these units turn on, we think they induce a magnetic field from
the inrush current that briefly disrupts the maser. We don't think it's
electrical because moving to another phase did not change things.
These air conditioners are all quite close to the masers. Typically a metre
or 2 away.
Much was done to discover this, but the clincher was that when the weather
cooled enough at the southern most location (Hobart), we turned off the air
con (only heating was needed) and the problem vanished.
So there's a lesson here for all maser owners. The jump of 0.7 nsec is not
much, but it's huge for VLBI and for time-nuts.
Jim Palfreyman
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Hi Jim,
Thanks much for the update. I can see how this was a pain to track down.
For those that don't remember the issue, the archive starts here:
https://www.febo.com/pipermail/time-nuts/2015-December/094904.html
And note that 1 / 0.704 ns = 1420 MHz, the frequency of a H-maser.
So it's either electrical or magnetic or seismic, yes? Does it happen every time the HVAC turns on/off? Can you run a high-resolution phase comparator during the event to find the time constant of the phase jump / cycle slip? 1 kHz sample rate should be more than enough. That may help narrow down which circuit is at fault.
/tvb
----- Original Message -----
From: "Jim Palfreyman" jim77742@gmail.com
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Saturday, May 21, 2016 6:58 PM
Subject: [time-nuts] Maser 0.7 nsec jumps solved
Hi all,
Awhile back I posted about some mysterious 0.7 ns jumps in three different
masers (of the same brand) at three different locations around Australia.
Well we think we've found the problem. All three locations also have
in-room air conditioners of the same brand. These are used for cooling
only. When these units turn on, we think they induce a magnetic field from
the inrush current that briefly disrupts the maser. We don't think it's
electrical because moving to another phase did not change things.
These air conditioners are all quite close to the masers. Typically a metre
or 2 away.
Much was done to discover this, but the clincher was that when the weather
cooled enough at the southern most location (Hobart), we turned off the air
con (only heating was needed) and the problem vanished.
So there's a lesson here for all maser owners. The jump of 0.7 nsec is not
much, but it's huge for VLBI and for time-nuts.
Jim Palfreyman
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Interesting math: Hydrogen maser frequency standards use the 1420 MHz line.
Period of 1420MHz is 0.7 ns.
It's not so clear to me that the maser itself is being disrupted, it seems
more likely the external noise is inducing an extra count or causing a
count to be slipped.
A different Australian observatory 1400 MHz RFI problem:
http://arxiv.org/abs/1504.02165 "Subsequent tests revealed that a peryton
can be generated at 1.4 GHz when a microwave oven door is opened
prematurely and the telescope is at an appropriate relative angle. Radio
emission escaping from microwave ovens during the magnetron shut-down phase
neatly explain all of the observed properties of the peryton signals."
Tim N3QE
On Sun, May 22, 2016 at 7:43 AM, Magnus Danielson <
magnus@rubidium.dyndns.org> wrote:
Jim,
On 05/22/2016 03:58 AM, Jim Palfreyman wrote:
Hi all,
Awhile back I posted about some mysterious 0.7 ns jumps in three different
masers (of the same brand) at three different locations around Australia.
Well we think we've found the problem. All three locations also have
in-room air conditioners of the same brand. These are used for cooling
only. When these units turn on, we think they induce a magnetic field from
the inrush current that briefly disrupts the maser. We don't think it's
electrical because moving to another phase did not change things.
These air conditioners are all quite close to the masers. Typically a
metre
or 2 away.
Much was done to discover this, but the clincher was that when the weather
cooled enough at the southern most location (Hobart), we turned off the
air
con (only heating was needed) and the problem vanished.
So there's a lesson here for all maser owners. The jump of 0.7 nsec is not
much, but it's huge for VLBI and for time-nuts.
Good that you have found the offender, but have you been able to remedy it
by other means than turning the AC off?
I think others H-maser owners would love to know, and potentially the
vendor you have.
Cheers,
Magnus
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
As far as a remedy goes we are going to try a solid state relay that only
switches on at 0V in the AC waveform. This should slow the inrush current,
and hopefully the magnetic impulse.
If this doesn't work, then a better model of air conditioner might have to
be installed. These ones do come on with a big "thump".
Jim Palfreyman
On 22 May 2016 at 21:43, Magnus Danielson magnus@rubidium.dyndns.org
wrote:
Jim,
On 05/22/2016 03:58 AM, Jim Palfreyman wrote:
Hi all,
Awhile back I posted about some mysterious 0.7 ns jumps in three different
masers (of the same brand) at three different locations around Australia.
Well we think we've found the problem. All three locations also have
in-room air conditioners of the same brand. These are used for cooling
only. When these units turn on, we think they induce a magnetic field from
the inrush current that briefly disrupts the maser. We don't think it's
electrical because moving to another phase did not change things.
These air conditioners are all quite close to the masers. Typically a
metre
or 2 away.
Much was done to discover this, but the clincher was that when the weather
cooled enough at the southern most location (Hobart), we turned off the
air
con (only heating was needed) and the problem vanished.
So there's a lesson here for all maser owners. The jump of 0.7 nsec is not
much, but it's huge for VLBI and for time-nuts.
Good that you have found the offender, but have you been able to remedy it
by other means than turning the AC off?
I think others H-maser owners would love to know, and potentially the
vendor you have.
Cheers,
Magnus
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
In message CALH-g5b0C+aBz53MUM7bkJ00441AvNxwFLc2RgjgPj=k1FPOKA@mail.gmail.com, Jim Palfr
eyman writes:
As far as a remedy goes we are going to try a solid state relay that only
switches on at 0V in the AC waveform. This should slow the inrush current,
and hopefully the magnetic impulse.
If that is not enough, consider a small VFD drive, and ramp up voltage+frequency
over 10 seconds. It's slightly more intrusive, as you will need to remove the
starting capacitor from the motor to install the VFD.
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
On Sun, May 22, 2016 at 11:15 PM, Jim Palfreyman jim77742@gmail.com wrote:
As far as a remedy goes we are going to try a solid state relay that only
switches on at 0V in the AC waveform. This should slow the inrush current,
and hopefully the magnetic impulse.
If the load being switched on is inductive, it would be better to switch
the AC waveform at the voltage peaks, not at 0V. This might seem
counter-intuitive, but it's real. Switching on at the 0V crossing may
maximize the current pulse through the magnetics.
OTOH, if the origin of the impulse is mechanical in nature, neither remedy
may help.
Regards,
Andy
Am 23.05.2016 um 05:15 schrieb Jim Palfreyman:
As far as a remedy goes we are going to try a solid state relay that only
switches on at 0V in the AC waveform. This should slow the inrush current,
and hopefully the magnetic impulse.
In the context of transformers and motors, switching on at 0V is
actually the worst point in time.
< https://de.wikipedia.org/wiki/Einschalten_des_Transformators > (German)
<
http://electrical-engineering-portal.com/practical-considerations-of-transformer-inrush-current
regards, Gerhard
In message 57441ACA.8070609@arcor.de, Gerhard Hoffmann writes:
Am 23.05.2016 um 05:15 schrieb Jim Palfreyman:
As far as a remedy goes we are going to try a solid state relay that only
switches on at 0V in the AC waveform. This should slow the inrush current,
and hopefully the magnetic impulse.
In the context of transformers and motors, switching on at 0V is
actually the worst point in time.
Well...
The motor is almost certainly an "Squirrel-cage" induction motor and
that means it is three phase, although one of the phases is probably
created with a "starting capacitor".
So which of the three phases is going to be the lucky one that
switches at maximum voltage, or are you going to switch the phases
on sequentially ?
A 4kW Variable Frequency Drive costs less than $1k and allows you
to control both the voltage/time and the frequency/time both
during startup and during rundown.
I wouldn't bother fuzzing around with hacks - I'd just go for the
known-to-work solution.
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
Am 23.05.2016 um 05:15 schrieb Jim Palfreyman:
As far as a remedy goes we are going to try a solid state relay that only
switches on at 0V in the AC waveform. This should slow the inrush current,
and hopefully the magnetic impulse.
In the context of transformers and motors, switching on at 0V is
actually the worst point in time.
< https://de.wikipedia.org/wiki/Einschalten_des_Transformators > (German)
<
http://electrical-engineering-portal.com/practical-considerations-of-transformer-inrush-current
regards, Gerhard
LTspice shows switching at 0V is the best point in time. With no
flux in the magnetics, the inrush current is limited by the circuit
resistance. The magnitude is given by Ohm's law: I = E / R.
Switching at 0V,
I = E / R
= 0 / R
= 0A
However, switching at the peak of the voltage can give very high
inrush currents. You can verify this with LTspice. The schematic is
at
http://www.pst.netii.net/misc/48b961a6.gif
There are two circuits. They are identical except the second has the
applied voltage shifted by 90 degrees. The inductors have 1 Ohm
series resistance (not shown.)
The waveforms are at
http://www.pst.netii.net/misc/48b96217.gif
The currents are IL1 and IL2.
Switching at 0V, IL1 starts at zero. It rises to a peak of 900mA,
then falls back to zero. The DC offset takes 3 or 4 seconds to
decay, then the current is stable at zero +/- 450mA.
Switching at the peak, IL2 is
I = E / R
= 169.7 / 1
= 169.7A
It takes over twice as long for the starting surge to decay. I could
still detect it past 8 seconds.
This analysis shows switching at 0V is the best option.
If you wish to do further analysis, the LTspice .ASC and .PLT files
are at
http://www.pst.netii.net/misc/48b96262.zip
The wikipedia article states "To avoid magnetic inrush, only for
transformers with an air gap in the core, the inductive load needs
to be synchronously connected near a supply voltage peak."
https://en.wikipedia.org/wiki/Inrush_current
Clearly, from the above LTspice waveforms, switching at the peak
gives the highest inrush surge that is possible to obtain. It also
takes the longest time to decay.
MRM
In message b222ff$74b01p@out.teksavvy.com, "Mike Monett" writes:
The wikipedia article states "To avoid magnetic inrush, only for
transformers with an air gap in the core, the inductive load needs
to be synchronously connected near a supply voltage peak."
You ought to have stopped and wondered at the "with an air gap in the core"
and pondered if you SPICE models use of an ideal inductor was appropriate
for something as electrically complex as an induction motor...
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
On Wed, May 25, 2016 at 9:59 AM, Mike Monett timenuts@binsamp.e4ward.com
wrote:
Am 23.05.2016 um 05:15 schrieb Jim Palfreyman:
As far as a remedy goes we are going to try a solid state relay that
only
switches on at 0V in the AC waveform. This should slow the inrush
current,
and hopefully the magnetic impulse.
In the context of transformers and motors, switching on at 0V is
actually the worst point in time.
< https://de.wikipedia.org/wiki/Einschalten_des_Transformators > (German)
<
regards, Gerhard
LTspice shows switching at 0V is the best point in time. With no
flux in the magnetics, the inrush current is limited by the circuit
resistance. The magnitude is given by Ohm's law: I = E / R.
Switching at 0V,
I = E / R
= 0 / R
= 0A
However, switching at the peak of the voltage can give very high
inrush currents. You can verify this with LTspice. The schematic is
at
http://www.pst.netii.net/misc/48b961a6.gif
There are two circuits. They are identical except the second has the
applied voltage shifted by 90 degrees. The inductors have 1 Ohm
series resistance (not shown.)
The waveforms are at
http://www.pst.netii.net/misc/48b96217.gif
The currents are IL1 and IL2.
Switching at 0V, IL1 starts at zero. It rises to a peak of 900mA,
then falls back to zero. The DC offset takes 3 or 4 seconds to
decay, then the current is stable at zero +/- 450mA.
Switching at the peak, IL2 is
I = E / R
= 169.7 / 1
= 169.7A
Er, no. It's an RL circuit for which:
V(R) = E(1-e^(-(R/L)t)
so I(R) = I(L) = (E/R)(1-e^(-(R/L)t)
for t close to zero where E is essentially constant, I is going to rise
with the time constant of R/L which for a 1H inductor and 1 ohm resistance
is 1 second.
I suggest running the simulation with a pulse voltage source, Tdelay 1ms,
period 1s, Ton 0.5s (Period and Ton are fairly arbitrary, the Tdelay is
important).
You only need to run the simulation for a few ms. You will get a very
different result.
Now as Poul-Henning Kamp suggested, an induction motor is anything but an
ideal inductor...
Am 25.05.2016 um 18:59 schrieb Mike Monett:
This analysis shows switching at 0V is the best option.
No, it doesn't. :-)
First, the single inductor does not represent a transformer; the second
inductor
and the coupling declaration ( style: K1 L1 L2 0.99 or so) and the load are
missing.
The most important thing is that the Inductor is nonlinear which is not
represented in the model. If there has remained some magnetism in
the core from previous operation, the transformer won't be able to further
increase the magnetism as needed to induce an opposing voltage
in the primary winding. When the core saturates, the inductance
collapses and leaves only the copper resistance to limit the current.
The catastrophe builds up in the first 90 degrees of the source wave,
not right at the start.
In the simulation, the current at t=0 is at its maximum already when at
t=0 the input voltage is just switched on. You'd expect that from a
capacitor, never from an inductor.
The reason is that the simulation does not really start at t=0, but
much earlier. The simulator computes the conductance matrix,
applies the initial sources and waits until everything has calmed down.
That may require repeated recalculation of the matrix to respect
nonlinearities.
Your circuit must contain hidden resistors btw, otherwise the computation
of the initial condition at "t<=0" would result in numeric overflow
as required by an assumed initial DC voltage across an inductor, before
the transient simulation.
One can enforce initial conditions with statements like .IC v(my_node) = 0V
regards, Gerhard
BTW I've got the first 20 pcs. of the VCXO carrier / voltage regulator /
lock to reference / squarer / iso amp or frequency doubler / 1pps board.
That won't be soldering for beginners or jittery hands. :-)
On Wed, May 25, 2016 at 12:59 PM, Mike Monett timenuts@binsamp.e4ward.com
wrote:
LTspice shows switching at 0V is the best point in time. ...
Bzzzt! Your simulation is seriously flawed, and your conclusions are
wrong. What you forgot, or may not have realized, is that SPICE's initial
transient solution is obtained by having the signal sources already turned
on (at the moment of the Big Bang) and set to their initial value, so the
current through L2 is limited by DC conditions. That is not anything close
to switching the driving voltages on. It is having one waveform sit at
+169.7V DC for a very long time ('forever'), and then letting it follow a
cosine wave.
Re-run the simulation with "UIC" added to the .tran statement (.tran 50ms
uic) and see what it shows. Using UIC forces the initial voltage to be 0V
at time=0, the start of the simulation. That's like having the switch
initially open.
Or if you don't like that, multiply the sources by a PWL waveform that
starts both voltages at 0V and then switches them on, a few milliseconds
into the simulation, with the appropriate phase.
Or use an actual switch. LTspice has a switch element you could use.
I guarantee you, the case with the voltage switching on at the 0V point in
the voltage waveform, causes greater currents.
The smaller surge current happens when the source is connected at the
moment when the current i(t) would be 0A if it were a continuous waveform.
For an inductive load, this happens when the voltage v(t) would be +/- peak
(or near peak, for a real load which has both inductance and a little
resistance). This condition also results in no surge, thus no L/R decay.
All of this might not be relevant to a mechanical system, where surge
current is caused by rotational inertia, rather than anything electrical.
Regards,
Andy
Hi All,
Thanks so much for your input and thoughts. It has really proved helpful
here at the observatory.
As it turned out we easily obtained a zero-crossing solid state relay so we
thought we'd try it.
And, drumroll......
It made things so much terribly worse than ever before. (As predicted by
many of you above.)
We are going to try a SSR that switches at the peak - but we need to order
one. So stay tuned on those results.
There is of course the "move the bloody thing far away from the maser"
solution which could end up being a serious option. These air conditioning
units are small and cheap (window-type), so we are trying to find the
cheapest solution - and if that ends up being some ducting - so be it!
Jim Palfreyman
On 26 May 2016 at 13:13, Andy AI.egrps+tn@gmail.com wrote:
On Wed, May 25, 2016 at 12:59 PM, Mike Monett <timenuts@binsamp.e4ward.com
wrote:
LTspice shows switching at 0V is the best point in time. ...
Bzzzt! Your simulation is seriously flawed, and your conclusions are
wrong. What you forgot, or may not have realized, is that SPICE's initial
transient solution is obtained by having the signal sources already turned
on (at the moment of the Big Bang) and set to their initial value, so the
current through L2 is limited by DC conditions. That is not anything close
to switching the driving voltages on. It is having one waveform sit at
+169.7V DC for a very long time ('forever'), and then letting it follow a
cosine wave.
Re-run the simulation with "UIC" added to the .tran statement (.tran 50ms
uic) and see what it shows. Using UIC forces the initial voltage to be 0V
at time=0, the start of the simulation. That's like having the switch
initially open.
Or if you don't like that, multiply the sources by a PWL waveform that
starts both voltages at 0V and then switches them on, a few milliseconds
into the simulation, with the appropriate phase.
Or use an actual switch. LTspice has a switch element you could use.
I guarantee you, the case with the voltage switching on at the 0V point in
the voltage waveform, causes greater currents.
The smaller surge current happens when the source is connected at the
moment when the current i(t) would be 0A if it were a continuous waveform.
For an inductive load, this happens when the voltage v(t) would be +/- peak
(or near peak, for a real load which has both inductance and a little
resistance). This condition also results in no surge, thus no L/R decay.
All of this might not be relevant to a mechanical system, where surge
current is caused by rotational inertia, rather than anything electrical.
Regards,
Andy
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Jim,
My head is precisely in the get it away from the unit approach.
Did not mention it for the following reason.
Its well understood and for time-nut boring. Its more fun to figure out
peak currents and such.
But I tend to fall into the get it done camp and move on.
That doesn't mean its a simple answer. Flex duct is bad. So only use it in
the last few feet. You want low resistance hard duct.
Then the fun of the return feed. Often overlooked and poorly considered.
Regards
Paul
WB8TSL
On Fri, Jun 3, 2016 at 1:00 AM, Jim Palfreyman jim77742@gmail.com wrote:
Hi All,
Thanks so much for your input and thoughts. It has really proved helpful
here at the observatory.
As it turned out we easily obtained a zero-crossing solid state relay so we
thought we'd try it.
And, drumroll......
It made things so much terribly worse than ever before. (As predicted by
many of you above.)
We are going to try a SSR that switches at the peak - but we need to order
one. So stay tuned on those results.
There is of course the "move the bloody thing far away from the maser"
solution which could end up being a serious option. These air conditioning
units are small and cheap (window-type), so we are trying to find the
cheapest solution - and if that ends up being some ducting - so be it!
Jim Palfreyman
On 26 May 2016 at 13:13, Andy AI.egrps+tn@gmail.com wrote:
On Wed, May 25, 2016 at 12:59 PM, Mike Monett <
wrote:
LTspice shows switching at 0V is the best point in time. ...
Bzzzt! Your simulation is seriously flawed, and your conclusions are
wrong. What you forgot, or may not have realized, is that SPICE's
initial
transient solution is obtained by having the signal sources already
turned
on (at the moment of the Big Bang) and set to their initial value, so the
current through L2 is limited by DC conditions. That is not anything
close
to switching the driving voltages on. It is having one waveform sit at
+169.7V DC for a very long time ('forever'), and then letting it follow a
cosine wave.
Re-run the simulation with "UIC" added to the .tran statement (.tran 50ms
uic) and see what it shows. Using UIC forces the initial voltage to be
0V
at time=0, the start of the simulation. That's like having the switch
initially open.
Or if you don't like that, multiply the sources by a PWL waveform that
starts both voltages at 0V and then switches them on, a few milliseconds
into the simulation, with the appropriate phase.
Or use an actual switch. LTspice has a switch element you could use.
I guarantee you, the case with the voltage switching on at the 0V point
in
the voltage waveform, causes greater currents.
The smaller surge current happens when the source is connected at the
moment when the current i(t) would be 0A if it were a continuous
waveform.
For an inductive load, this happens when the voltage v(t) would be +/-
peak
(or near peak, for a real load which has both inductance and a little
resistance). This condition also results in no surge, thus no L/R decay.
All of this might not be relevant to a mechanical system, where surge
current is caused by rotational inertia, rather than anything electrical.
Regards,
Andy
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
One more thought.
Is the aircon on the same power phase as the maser?
Are you split phase in the facility at least.
On Fri, Jun 3, 2016 at 9:24 AM, paul swed paulswedb@gmail.com wrote:
Jim,
My head is precisely in the get it away from the unit approach.
Did not mention it for the following reason.
Its well understood and for time-nut boring. Its more fun to figure out
peak currents and such.
But I tend to fall into the get it done camp and move on.
That doesn't mean its a simple answer. Flex duct is bad. So only use it in
the last few feet. You want low resistance hard duct.
Then the fun of the return feed. Often overlooked and poorly considered.
Regards
Paul
WB8TSL
On Fri, Jun 3, 2016 at 1:00 AM, Jim Palfreyman jim77742@gmail.com wrote:
Hi All,
Thanks so much for your input and thoughts. It has really proved helpful
here at the observatory.
As it turned out we easily obtained a zero-crossing solid state relay so
we
thought we'd try it.
And, drumroll......
It made things so much terribly worse than ever before. (As predicted by
many of you above.)
We are going to try a SSR that switches at the peak - but we need to order
one. So stay tuned on those results.
There is of course the "move the bloody thing far away from the maser"
solution which could end up being a serious option. These air conditioning
units are small and cheap (window-type), so we are trying to find the
cheapest solution - and if that ends up being some ducting - so be it!
Jim Palfreyman
On 26 May 2016 at 13:13, Andy AI.egrps+tn@gmail.com wrote:
On Wed, May 25, 2016 at 12:59 PM, Mike Monett <
wrote:
LTspice shows switching at 0V is the best point in time. ...
Bzzzt! Your simulation is seriously flawed, and your conclusions are
wrong. What you forgot, or may not have realized, is that SPICE's
initial
transient solution is obtained by having the signal sources already
turned
on (at the moment of the Big Bang) and set to their initial value, so
the
current through L2 is limited by DC conditions. That is not anything
close
to switching the driving voltages on. It is having one waveform sit at
+169.7V DC for a very long time ('forever'), and then letting it follow
a
cosine wave.
Re-run the simulation with "UIC" added to the .tran statement (.tran
50ms
uic) and see what it shows. Using UIC forces the initial voltage to be
0V
at time=0, the start of the simulation. That's like having the switch
initially open.
Or if you don't like that, multiply the sources by a PWL waveform that
starts both voltages at 0V and then switches them on, a few milliseconds
into the simulation, with the appropriate phase.
Or use an actual switch. LTspice has a switch element you could use.
I guarantee you, the case with the voltage switching on at the 0V point
in
the voltage waveform, causes greater currents.
The smaller surge current happens when the source is connected at the
moment when the current i(t) would be 0A if it were a continuous
waveform.
For an inductive load, this happens when the voltage v(t) would be +/-
peak
(or near peak, for a real load which has both inductance and a little
resistance). This condition also results in no surge, thus no L/R
decay.
All of this might not be relevant to a mechanical system, where surge
current is caused by rotational inertia, rather than anything
electrical.
Regards,
Andy
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
To All;
I found a significant error in the LTspice analysis. I was wondering how
the current could jump instantaneously at zero when the voltage is applied
at the peak. That violates magnetism.
It turns out it doesn't. When LTspice starts an analysis, it first
calculates the operating point. For the Sine voltage source at 90
degrees, it applies the full voltage across the load. In this case,
it was 169.7V across 1 ohm, resulting in 169.7 Amps. That is what
was plotted, and is a significant error.
I embarked on a search to find examples where switching at the peak could
reduce the inrush current.
Out of 13 examples I analyzed, I found only one that involves unloaded
transformers.
I found many references that discuss transformer inrush current caused by
core saturation. This is a serious problem as it puts stress on the
components and reduces operating life.
I could find no reference that states switching at the peak would reduce or
eliminate the inrush current. I also found most major suppliers do not
offer SSR's that will switch at the peak.
Obviously, switching at the peak would be worse for capacitive loads.
This was a major project and turned out to take a lot more time and effort
than expected. For those who may be interested, the results are shown at
http://www.pst.netii.net/timenuts/zvs.htm
MRM
On Fri, 03 Jun 2016 12:37:26 -0400
"Mike Monett" timenuts@binsamp.e4ward.com wrote:
I found a significant error in the LTspice analysis. I was wondering how
the current could jump instantaneously at zero when the voltage is applied
at the peak. That violates magnetism.
It turns out it doesn't. When LTspice starts an analysis, it first
calculates the operating point. For the Sine voltage source at 90
degrees, it applies the full voltage across the load. In this case,
it was 169.7V across 1 ohm, resulting in 169.7 Amps. That is what
was plotted, and is a significant error.
Actually, spice (the engine behind LTspice) does a DC analysis before
almost all modes of operation. This DC analysis has the intention to
start the circuit from a steady-state point and thus to reduce simulation
time. In order for this to work properly, you have to specify the DC voltage
and currents for all sources correctly. Spice messes this up at times
making the first part of a transient simulation worthless (it has even
worse problems when you do an AC analysis). Additionally LTspice hides
too much of these small complications for the problems to be visible to
the untrained eye and also at times makes it harder to provide the correct
values. Thus, caution is advised.
The general rule of "Never trust a simulation you haven't
forged yourself" applies.
Out of 13 examples I analyzed, I found only one that involves unloaded
transformers.
I found many references that discuss transformer inrush current caused by
core saturation. This is a serious problem as it puts stress on the
components and reduces operating life.
I only had a quick glance at your webpage, but it seems that you used
the standard LTspice transformer model. Unfortunately, this is not a
good model to study this kind of behaviour. For one, the only loss considered
in the model is the winding coupling, it doesn't even directly consider
resistive losses in the windings. In this case, the two most important effects
that you need to include are saturation and core losses, which are both
frequency dependent. The cores of electric machine transformers are very
poor when it comes to their "high" frequency behaviour. Where high frequency
starts somewhere closely above mains frequency. Ie 1kHz is already so far off
that somewhere around 90% of the energy would be dissipated in the core.
The sharp rise in voltage and the leading inrush current have frequency
components that are way higher than mains frequency. Hence the linear model
you used will give inaccurate results, to put it mildly.
Unfortunately, building an accurate transformer model in spice is not
easy and depends on higher order functions that might or might not be
available in the flavour you use. Not to mention that you will need
to have good (measured) numbers on the non-ideal behaviour of a transformer,
which are also not easy to get by.
Attila Kinali
--
It is upon moral qualities that a society is ultimately founded. All
the prosperity and technological sophistication in the world is of no
use without that foundation.
-- Miss Matheson, The Diamond Age, Neil Stephenson