Re: [time-nuts] Question for the cesium nuts.

R
rlutwak@comcast.net
Fri, Mar 16, 2007 5:55 PM

What are we going to do when all the CBTs owned by amateurs start to end of
life?

Buy another instrument off of Ebay.  It'll be cheaper, more accurate, and last longer than the old one.  Plus, it'll have microprocessor control and thus be cooler and more entertaining for the hackers.

Anyone care to start a discussion on the merits of restoring a CBT to life?

It can't be done.  Trust me, I've done it.

-RL

--

Robert Lutwak, Senior Scientist
Symmetricom - Technology Realization Center
34 Tozer Rd.
Beverly, MA 01915

RLutwak@Symmetricom.com (Business)
Lutwak@Alum.MIT.edu (Personal)

(978) 232-1461 (Desk)
(339) 927-7896 (Mobile)
(978) 927-4099 (FAX)

-------------- Original message --------------
From: "Jack Hudler" jack@hudler.org

It seems to me that like all good things they must come to and end.

If all CBTs have a life expectancy that varies depending on the
manufacturer.

What are we going to do when all the CBTs owned by amateurs start to end of
life?

I for one am certainly not going to buy one, not at those prices! (Unless
I'm retired then that's another story)
You only have calculate the time value of money for that CBT purchase over
the remaining time to retirement; If that doesn't stop you dead in your
tracks then this group really is aptly named! :)

From my perspective, that of wanting to own a Cesium Standard; I don't
really want to layout the monies for something that's going to end of life
on me shortly (few years) afterwards.

I know that handling (Caesium) Cesium-133 is tricky at best. It's a heavy
alkali metal and contact with moisture is right out!
Other than that it's not terribly difficult to create a safe environment to
work with it.

So there must be something else that's considerably more difficult than
opening the tube, recharging the ampoule, resealing it, pulling an ultra
high vacuum and baking it out.

I've not seen any pictures of a naked CBT, still I'm not too worried about
cracking the tube open if its Pyrex, unless resealing it caused the cesium
beam collimation to be lost.

Are there if any getters to worry about? If so, how would one ablate the
contaminates of the surface?

Anyone care to start a discussion on the merits of restoring a CBT to life?

Jack


time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

> What are we going to do when all the CBTs owned by amateurs start to end of > life? Buy another instrument off of Ebay. It'll be cheaper, more accurate, and last longer than the old one. Plus, it'll have microprocessor control and thus be cooler and more entertaining for the hackers. > Anyone care to start a discussion on the merits of restoring a CBT to life? It can't be done. Trust me, I've done it. -RL -- ---------------- Robert Lutwak, Senior Scientist Symmetricom - Technology Realization Center 34 Tozer Rd. Beverly, MA 01915 RLutwak@Symmetricom.com (Business) Lutwak@Alum.MIT.edu (Personal) (978) 232-1461 (Desk) (339) 927-7896 (Mobile) (978) 927-4099 (FAX) -------------- Original message -------------- From: "Jack Hudler" <jack@hudler.org> > It seems to me that like all good things they must come to and end. > > If all CBTs have a life expectancy that varies depending on the > manufacturer. > > What are we going to do when all the CBTs owned by amateurs start to end of > life? > > I for one am certainly not going to buy one, not at those prices! (Unless > I'm retired then that's another story) > You only have calculate the time value of money for that CBT purchase over > the remaining time to retirement; If that doesn't stop you dead in your > tracks then this group really is aptly named! :) > > From my perspective, that of wanting to own a Cesium Standard; I don't > really want to layout the monies for something that's going to end of life > on me shortly (few years) afterwards. > > I know that handling (Caesium) Cesium-133 is tricky at best. It's a heavy > alkali metal and contact with moisture is right out! > Other than that it's not terribly difficult to create a safe environment to > work with it. > > So there must be something else that's considerably more difficult than > opening the tube, recharging the ampoule, resealing it, pulling an ultra > high vacuum and baking it out. > > I've not seen any pictures of a naked CBT, still I'm not too worried about > cracking the tube open if its Pyrex, unless resealing it caused the cesium > beam collimation to be lost. > > Are there if any getters to worry about? If so, how would one ablate the > contaminates of the surface? > > Anyone care to start a discussion on the merits of restoring a CBT to life? > > Jack > > > > > > _______________________________________________ > time-nuts mailing list > time-nuts@febo.com > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
JH
Jack Hudler
Fri, Mar 16, 2007 6:25 PM

Which begs the question of; who's cesium standard will we buy in surplus
market?

Besides you guys (Symmetricom), who is building Cesium standards that
haven't yet been absorbed by Symmetricom? (note this is not meant to be
derogatory).

Certainly most if not all the 5060, 5061, and 5062 are either dead or close
to it. Excluding those that have had their CBT's replaced or properly stored
and regularly pumped down, or just lucky.

The 5071's have yet to make an appearance on eBay at levels I would consider
paying.

From my POV (which could be myopic), a few CBT manufacturers are controlling

what remains of this market (no I'm not a conspiracy nut, it's just
business), so it seems to me that the surplus market is going to get very
thin in the near future.
Supply and demand dictates that surplus market prices will skyrocket out of
the vast majority of amateur reaches in the coming years.

So what's the next cesium standard to start showing up on eBay in numbers
with life left?

Jack

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of rlutwak@comcast.net
Sent: Friday, March 16, 2007 12:56 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Question for the cesium nuts.

What are we going to do when all the CBTs owned by amateurs start to end

of

life?

Buy another instrument off of Ebay.  It'll be cheaper, more accurate, and
last longer than the old one.  Plus, it'll have microprocessor control and
thus be cooler and more entertaining for the hackers.

Anyone care to start a discussion on the merits of restoring a CBT to

life?
It can't be done.  Trust me, I've done it.

-RL

--

Robert Lutwak, Senior Scientist
Symmetricom - Technology Realization Center
34 Tozer Rd.
Beverly, MA 01915

RLutwak@Symmetricom.com (Business)
Lutwak@Alum.MIT.edu (Personal)

(978) 232-1461 (Desk)
(339) 927-7896 (Mobile)
(978) 927-4099 (FAX)

-------------- Original message --------------
From: "Jack Hudler" jack@hudler.org

It seems to me that like all good things they must come to and end.

If all CBTs have a life expectancy that varies depending on the
manufacturer.

What are we going to do when all the CBTs owned by amateurs start to end

of

life?

I for one am certainly not going to buy one, not at those prices! (Unless
I'm retired then that's another story)
You only have calculate the time value of money for that CBT purchase over

the remaining time to retirement; If that doesn't stop you dead in your
tracks then this group really is aptly named! :)

From my perspective, that of wanting to own a Cesium Standard; I don't
really want to layout the monies for something that's going to end of life

on me shortly (few years) afterwards.

I know that handling (Caesium) Cesium-133 is tricky at best. It's a heavy
alkali metal and contact with moisture is right out!
Other than that it's not terribly difficult to create a safe environment

to

work with it.

So there must be something else that's considerably more difficult than
opening the tube, recharging the ampoule, resealing it, pulling an ultra
high vacuum and baking it out.

I've not seen any pictures of a naked CBT, still I'm not too worried about

cracking the tube open if its Pyrex, unless resealing it caused the cesium

beam collimation to be lost.

Are there if any getters to worry about? If so, how would one ablate the
contaminates of the surface?

Anyone care to start a discussion on the merits of restoring a CBT to

life?

Which begs the question of; who's cesium standard will we buy in surplus market? Besides you guys (Symmetricom), who is building Cesium standards that haven't yet been absorbed by Symmetricom? (note this is not meant to be derogatory). Certainly most if not all the 5060, 5061, and 5062 are either dead or close to it. Excluding those that have had their CBT's replaced or properly stored and regularly pumped down, or just lucky. The 5071's have yet to make an appearance on eBay at levels I would consider paying. >From my POV (which could be myopic), a few CBT manufacturers are controlling what remains of this market (no I'm not a conspiracy nut, it's just business), so it seems to me that the surplus market is going to get very thin in the near future. Supply and demand dictates that surplus market prices will skyrocket out of the vast majority of amateur reaches in the coming years. So what's the next cesium standard to start showing up on eBay in numbers with life left? Jack -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On Behalf Of rlutwak@comcast.net Sent: Friday, March 16, 2007 12:56 PM To: Discussion of precise time and frequency measurement Subject: Re: [time-nuts] Question for the cesium nuts. > What are we going to do when all the CBTs owned by amateurs start to end of > life? Buy another instrument off of Ebay. It'll be cheaper, more accurate, and last longer than the old one. Plus, it'll have microprocessor control and thus be cooler and more entertaining for the hackers. > Anyone care to start a discussion on the merits of restoring a CBT to life? It can't be done. Trust me, I've done it. -RL -- ---------------- Robert Lutwak, Senior Scientist Symmetricom - Technology Realization Center 34 Tozer Rd. Beverly, MA 01915 RLutwak@Symmetricom.com (Business) Lutwak@Alum.MIT.edu (Personal) (978) 232-1461 (Desk) (339) 927-7896 (Mobile) (978) 927-4099 (FAX) -------------- Original message -------------- From: "Jack Hudler" <jack@hudler.org> > It seems to me that like all good things they must come to and end. > > If all CBTs have a life expectancy that varies depending on the > manufacturer. > > What are we going to do when all the CBTs owned by amateurs start to end of > life? > > I for one am certainly not going to buy one, not at those prices! (Unless > I'm retired then that's another story) > You only have calculate the time value of money for that CBT purchase over > the remaining time to retirement; If that doesn't stop you dead in your > tracks then this group really is aptly named! :) > > From my perspective, that of wanting to own a Cesium Standard; I don't > really want to layout the monies for something that's going to end of life > on me shortly (few years) afterwards. > > I know that handling (Caesium) Cesium-133 is tricky at best. It's a heavy > alkali metal and contact with moisture is right out! > Other than that it's not terribly difficult to create a safe environment to > work with it. > > So there must be something else that's considerably more difficult than > opening the tube, recharging the ampoule, resealing it, pulling an ultra > high vacuum and baking it out. > > I've not seen any pictures of a naked CBT, still I'm not too worried about > cracking the tube open if its Pyrex, unless resealing it caused the cesium > beam collimation to be lost. > > Are there if any getters to worry about? If so, how would one ablate the > contaminates of the surface? > > Anyone care to start a discussion on the merits of restoring a CBT to life? > > Jack > > > > > > _______________________________________________ > time-nuts mailing list > time-nuts@febo.com > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts _______________________________________________ time-nuts mailing list time-nuts@febo.com https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
MD
Magnus Danielson
Fri, Mar 16, 2007 7:36 PM

From: "Jack Hudler" jack@hudler.org
Subject: Re: [time-nuts] Question for the cesium nuts.
Date: Fri, 16 Mar 2007 13:25:47 -0500
Message-ID: 00ca01c767f8$84d91cc0$8e8b5640$@org

Which begs the question of; who's cesium standard will we buy in surplus
market?

Besides you guys (Symmetricom), who is building Cesium standards that
haven't yet been absorbed by Symmetricom? (note this is not meant to be
derogatory).

Oscilloquartz. I beleive that they have Symmetricom tubes in them.

Actually that would raise questions regarding to second-source for serious
telecom operators. Fortunatly for them, GPS clocks is usually enought.

I beleive there was a Russian Cesium-maker, but if I recall correctly they
stopped doing Cesium clocks.

From my POV (which could be myopic), a few CBT manufacturers are controlling

what remains of this market (no I'm not a conspiracy nut, it's just
business), so it seems to me that the surplus market is going to get very
thin in the near future.

It is really just buisness decissions behind it. Weither they where wise or not
is another issue. Joint buissnesses is not always beneficial (see second-source
comment).

Cheers,
Magnus

From: "Jack Hudler" <jack@hudler.org> Subject: Re: [time-nuts] Question for the cesium nuts. Date: Fri, 16 Mar 2007 13:25:47 -0500 Message-ID: <00ca01c767f8$84d91cc0$8e8b5640$@org> > Which begs the question of; who's cesium standard will we buy in surplus > market? > > Besides you guys (Symmetricom), who is building Cesium standards that > haven't yet been absorbed by Symmetricom? (note this is not meant to be > derogatory). Oscilloquartz. I beleive that they have Symmetricom tubes in them. Actually that would raise questions regarding to second-source for serious telecom operators. Fortunatly for them, GPS clocks is usually enought. I beleive there was a Russian Cesium-maker, but if I recall correctly they stopped doing Cesium clocks. > >From my POV (which could be myopic), a few CBT manufacturers are controlling > what remains of this market (no I'm not a conspiracy nut, it's just > business), so it seems to me that the surplus market is going to get very > thin in the near future. It is really just buisness decissions behind it. Weither they where wise or not is another issue. Joint buissnesses is not always beneficial (see second-source comment). Cheers, Magnus
RK
Rob Kimberley
Fri, Mar 16, 2007 8:21 PM

The only people I know making Cs now are Symmetricom (the old Datum/FTS +
recent Agilent acquisition), FEI (although they don't think they make many
now), and Temex in France. Heard rumours of some being made in Japan, but
not sure if and who. The Russians seem to concentrate on H2 Masers (Kvarz,
and Vremya).

Anyone know of others??

Rob K

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Jack Hudler
Sent: 16 March 2007 18:26
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] Question for the cesium nuts.

Which begs the question of; who's cesium standard will we buy in surplus
market?

Besides you guys (Symmetricom), who is building Cesium standards that
haven't yet been absorbed by Symmetricom? (note this is not meant to be
derogatory).

Certainly most if not all the 5060, 5061, and 5062 are either dead or close
to it. Excluding those that have had their CBT's replaced or properly stored
and regularly pumped down, or just lucky.

The 5071's have yet to make an appearance on eBay at levels I would consider
paying.

From my POV (which could be myopic), a few CBT manufacturers are controlling

what remains of this market (no I'm not a conspiracy nut, it's just
business), so it seems to me that the surplus market is going to get very
thin in the near future.
Supply and demand dictates that surplus market prices will skyrocket out of
the vast majority of amateur reaches in the coming years.

So what's the next cesium standard to start showing up on eBay in numbers
with life left?

Jack

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of rlutwak@comcast.net
Sent: Friday, March 16, 2007 12:56 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Question for the cesium nuts.

What are we going to do when all the CBTs owned by amateurs start to
end

of

life?

Buy another instrument off of Ebay.  It'll be cheaper, more accurate, and
last longer than the old one.  Plus, it'll have microprocessor control and
thus be cooler and more entertaining for the hackers.

Anyone care to start a discussion on the merits of restoring a CBT to

life?
It can't be done.  Trust me, I've done it.

-RL

--

Robert Lutwak, Senior Scientist
Symmetricom - Technology Realization Center
34 Tozer Rd.
Beverly, MA 01915

RLutwak@Symmetricom.com (Business)
Lutwak@Alum.MIT.edu (Personal)

(978) 232-1461 (Desk)
(339) 927-7896 (Mobile)
(978) 927-4099 (FAX)

-------------- Original message --------------
From: "Jack Hudler" jack@hudler.org

It seems to me that like all good things they must come to and end.

If all CBTs have a life expectancy that varies depending on the
manufacturer.

What are we going to do when all the CBTs owned by amateurs start to
end

of

life?

I for one am certainly not going to buy one, not at those prices!
(Unless I'm retired then that's another story) You only have calculate
the time value of money for that CBT purchase over

the remaining time to retirement; If that doesn't stop you dead in
your tracks then this group really is aptly named! :)

From my perspective, that of wanting to own a Cesium Standard; I don't
really want to layout the monies for something that's going to end of
life

on me shortly (few years) afterwards.

I know that handling (Caesium) Cesium-133 is tricky at best. It's a
heavy alkali metal and contact with moisture is right out!
Other than that it's not terribly difficult to create a safe
environment

to

work with it.

So there must be something else that's considerably more difficult
than opening the tube, recharging the ampoule, resealing it, pulling
an ultra high vacuum and baking it out.

I've not seen any pictures of a naked CBT, still I'm not too worried
about

cracking the tube open if its Pyrex, unless resealing it caused the
cesium

beam collimation to be lost.

Are there if any getters to worry about? If so, how would one ablate
the contaminates of the surface?

Anyone care to start a discussion on the merits of restoring a CBT to

life?

The only people I know making Cs now are Symmetricom (the old Datum/FTS + recent Agilent acquisition), FEI (although they don't think they make many now), and Temex in France. Heard rumours of some being made in Japan, but not sure if and who. The Russians seem to concentrate on H2 Masers (Kvarz, and Vremya). Anyone know of others?? Rob K -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On Behalf Of Jack Hudler Sent: 16 March 2007 18:26 To: 'Discussion of precise time and frequency measurement' Subject: Re: [time-nuts] Question for the cesium nuts. Which begs the question of; who's cesium standard will we buy in surplus market? Besides you guys (Symmetricom), who is building Cesium standards that haven't yet been absorbed by Symmetricom? (note this is not meant to be derogatory). Certainly most if not all the 5060, 5061, and 5062 are either dead or close to it. Excluding those that have had their CBT's replaced or properly stored and regularly pumped down, or just lucky. The 5071's have yet to make an appearance on eBay at levels I would consider paying. >From my POV (which could be myopic), a few CBT manufacturers are controlling what remains of this market (no I'm not a conspiracy nut, it's just business), so it seems to me that the surplus market is going to get very thin in the near future. Supply and demand dictates that surplus market prices will skyrocket out of the vast majority of amateur reaches in the coming years. So what's the next cesium standard to start showing up on eBay in numbers with life left? Jack -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On Behalf Of rlutwak@comcast.net Sent: Friday, March 16, 2007 12:56 PM To: Discussion of precise time and frequency measurement Subject: Re: [time-nuts] Question for the cesium nuts. > What are we going to do when all the CBTs owned by amateurs start to > end of > life? Buy another instrument off of Ebay. It'll be cheaper, more accurate, and last longer than the old one. Plus, it'll have microprocessor control and thus be cooler and more entertaining for the hackers. > Anyone care to start a discussion on the merits of restoring a CBT to life? It can't be done. Trust me, I've done it. -RL -- ---------------- Robert Lutwak, Senior Scientist Symmetricom - Technology Realization Center 34 Tozer Rd. Beverly, MA 01915 RLutwak@Symmetricom.com (Business) Lutwak@Alum.MIT.edu (Personal) (978) 232-1461 (Desk) (339) 927-7896 (Mobile) (978) 927-4099 (FAX) -------------- Original message -------------- From: "Jack Hudler" <jack@hudler.org> > It seems to me that like all good things they must come to and end. > > If all CBTs have a life expectancy that varies depending on the > manufacturer. > > What are we going to do when all the CBTs owned by amateurs start to > end of > life? > > I for one am certainly not going to buy one, not at those prices! > (Unless I'm retired then that's another story) You only have calculate > the time value of money for that CBT purchase over > the remaining time to retirement; If that doesn't stop you dead in > your tracks then this group really is aptly named! :) > > From my perspective, that of wanting to own a Cesium Standard; I don't > really want to layout the monies for something that's going to end of > life > on me shortly (few years) afterwards. > > I know that handling (Caesium) Cesium-133 is tricky at best. It's a > heavy alkali metal and contact with moisture is right out! > Other than that it's not terribly difficult to create a safe > environment to > work with it. > > So there must be something else that's considerably more difficult > than opening the tube, recharging the ampoule, resealing it, pulling > an ultra high vacuum and baking it out. > > I've not seen any pictures of a naked CBT, still I'm not too worried > about > cracking the tube open if its Pyrex, unless resealing it caused the > cesium > beam collimation to be lost. > > Are there if any getters to worry about? If so, how would one ablate > the contaminates of the surface? > > Anyone care to start a discussion on the merits of restoring a CBT to life? > > Jack > > > > > > _______________________________________________ > time-nuts mailing list > time-nuts@febo.com > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts _______________________________________________ time-nuts mailing list time-nuts@febo.com https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts _______________________________________________ time-nuts mailing list time-nuts@febo.com https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
RL
Robert Lutwak
Fri, Mar 16, 2007 10:09 PM

There's no reason that you shouldn't see just as many CsIIIs, Cs4000s, and
5071s on the used market tomorrow as you see 4040s and 506Xs today, just as
soon as they trickle down through the same channels as the older units did.

So far, of the modern cesium standards, I've only seen a few 5071s on the
used market, and they're still priced pretty high, but I'm certain the
others will trickle down before long.

Eventually, they'll be dirt cheap, because a new CsIII costs half of what a
4040 did when new.  They'll be more accurate, for reasons that you can read
about in 15 years of FCS proceedings, and they'll be more reliable because
of 15 years of electronics evolution and manufacturing improvements.

Oh, and they'll be more fun, because you'll be able to run Monitor3.  :-)

-RL

----- Original Message -----
From: "Jack Hudler" jack@hudler.org
To: "'Discussion of precise time and frequency measurement'"
time-nuts@febo.com
Sent: Friday, March 16, 2007 2:25 PM
Subject: Re: [time-nuts] Question for the cesium nuts.

Which begs the question of; who's cesium standard will we buy in surplus
market?

Besides you guys (Symmetricom), who is building Cesium standards that
haven't yet been absorbed by Symmetricom? (note this is not meant to be
derogatory).

Certainly most if not all the 5060, 5061, and 5062 are either dead or
close
to it. Excluding those that have had their CBT's replaced or properly
stored
and regularly pumped down, or just lucky.

The 5071's have yet to make an appearance on eBay at levels I would
consider
paying.

From my POV (which could be myopic), a few CBT manufacturers are
controlling
what remains of this market (no I'm not a conspiracy nut, it's just
business), so it seems to me that the surplus market is going to get very
thin in the near future.
Supply and demand dictates that surplus market prices will skyrocket out
of
the vast majority of amateur reaches in the coming years.

So what's the next cesium standard to start showing up on eBay in numbers
with life left?

Jack

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of rlutwak@comcast.net
Sent: Friday, March 16, 2007 12:56 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Question for the cesium nuts.

What are we going to do when all the CBTs owned by amateurs start to end

of

life?

Buy another instrument off of Ebay.  It'll be cheaper, more accurate, and
last longer than the old one.  Plus, it'll have microprocessor control and
thus be cooler and more entertaining for the hackers.

Anyone care to start a discussion on the merits of restoring a CBT to

life?
It can't be done.  Trust me, I've done it.

-RL

-------------- Original message --------------
From: "Jack Hudler" jack@hudler.org

It seems to me that like all good things they must come to and end.

If all CBTs have a life expectancy that varies depending on the
manufacturer.

What are we going to do when all the CBTs owned by amateurs start to end

of

life?

I for one am certainly not going to buy one, not at those prices! (Unless
I'm retired then that's another story)
You only have calculate the time value of money for that CBT purchase
over

the remaining time to retirement; If that doesn't stop you dead in your
tracks then this group really is aptly named! :)

From my perspective, that of wanting to own a Cesium Standard; I don't
really want to layout the monies for something that's going to end of
life

on me shortly (few years) afterwards.

I know that handling (Caesium) Cesium-133 is tricky at best. It's a heavy
alkali metal and contact with moisture is right out!
Other than that it's not terribly difficult to create a safe environment

to

work with it.

So there must be something else that's considerably more difficult than
opening the tube, recharging the ampoule, resealing it, pulling an ultra
high vacuum and baking it out.

I've not seen any pictures of a naked CBT, still I'm not too worried
about

cracking the tube open if its Pyrex, unless resealing it caused the
cesium

beam collimation to be lost.

Are there if any getters to worry about? If so, how would one ablate the
contaminates of the surface?

Anyone care to start a discussion on the merits of restoring a CBT to

life?

There's no reason that you shouldn't see just as many CsIIIs, Cs4000s, and 5071s on the used market tomorrow as you see 4040s and 506Xs today, just as soon as they trickle down through the same channels as the older units did. So far, of the modern cesium standards, I've only seen a few 5071s on the used market, and they're still priced pretty high, but I'm certain the others will trickle down before long. Eventually, they'll be dirt cheap, because a new CsIII costs half of what a 4040 did when new. They'll be more accurate, for reasons that you can read about in 15 years of FCS proceedings, and they'll be more reliable because of 15 years of electronics evolution and manufacturing improvements. Oh, and they'll be more fun, because you'll be able to run Monitor3. :-) -RL ----- Original Message ----- From: "Jack Hudler" <jack@hudler.org> To: "'Discussion of precise time and frequency measurement'" <time-nuts@febo.com> Sent: Friday, March 16, 2007 2:25 PM Subject: Re: [time-nuts] Question for the cesium nuts. > Which begs the question of; who's cesium standard will we buy in surplus > market? > > Besides you guys (Symmetricom), who is building Cesium standards that > haven't yet been absorbed by Symmetricom? (note this is not meant to be > derogatory). > > Certainly most if not all the 5060, 5061, and 5062 are either dead or > close > to it. Excluding those that have had their CBT's replaced or properly > stored > and regularly pumped down, or just lucky. > > The 5071's have yet to make an appearance on eBay at levels I would > consider > paying. > > From my POV (which could be myopic), a few CBT manufacturers are > controlling > what remains of this market (no I'm not a conspiracy nut, it's just > business), so it seems to me that the surplus market is going to get very > thin in the near future. > Supply and demand dictates that surplus market prices will skyrocket out > of > the vast majority of amateur reaches in the coming years. > > So what's the next cesium standard to start showing up on eBay in numbers > with life left? > > Jack > > -----Original Message----- > From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On > Behalf Of rlutwak@comcast.net > Sent: Friday, March 16, 2007 12:56 PM > To: Discussion of precise time and frequency measurement > Subject: Re: [time-nuts] Question for the cesium nuts. > >> What are we going to do when all the CBTs owned by amateurs start to end > of >> life? > Buy another instrument off of Ebay. It'll be cheaper, more accurate, and > last longer than the old one. Plus, it'll have microprocessor control and > thus be cooler and more entertaining for the hackers. > >> Anyone care to start a discussion on the merits of restoring a CBT to > life? > It can't be done. Trust me, I've done it. > > -RL > > > -------------- Original message -------------- > From: "Jack Hudler" <jack@hudler.org> > >> It seems to me that like all good things they must come to and end. >> >> If all CBTs have a life expectancy that varies depending on the >> manufacturer. >> >> What are we going to do when all the CBTs owned by amateurs start to end > of >> life? >> >> I for one am certainly not going to buy one, not at those prices! (Unless >> I'm retired then that's another story) >> You only have calculate the time value of money for that CBT purchase >> over > >> the remaining time to retirement; If that doesn't stop you dead in your >> tracks then this group really is aptly named! :) >> >> From my perspective, that of wanting to own a Cesium Standard; I don't >> really want to layout the monies for something that's going to end of >> life > >> on me shortly (few years) afterwards. >> >> I know that handling (Caesium) Cesium-133 is tricky at best. It's a heavy >> alkali metal and contact with moisture is right out! >> Other than that it's not terribly difficult to create a safe environment > to >> work with it. >> >> So there must be something else that's considerably more difficult than >> opening the tube, recharging the ampoule, resealing it, pulling an ultra >> high vacuum and baking it out. >> >> I've not seen any pictures of a naked CBT, still I'm not too worried >> about > >> cracking the tube open if its Pyrex, unless resealing it caused the >> cesium > >> beam collimation to be lost. >> >> Are there if any getters to worry about? If so, how would one ablate the >> contaminates of the surface? >> >> Anyone care to start a discussion on the merits of restoring a CBT to > life? >> >> Jack >> >> >> >> >> >> _______________________________________________ >> time-nuts mailing list >> time-nuts@febo.com >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > _______________________________________________ > time-nuts mailing list > time-nuts@febo.com > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > > _______________________________________________ > time-nuts mailing list > time-nuts@febo.com > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
MD
Magnus Danielson
Fri, Mar 16, 2007 10:29 PM

From: "Robert Lutwak" RLutwak@Comcast.net
Subject: Re: [time-nuts] Question for the cesium nuts.
Date: Fri, 16 Mar 2007 18:09:43 -0400
Message-ID: 001301c76817$cd143aa0$6700a8c0@symmetricom.com

Robert,

There's no reason that you shouldn't see just as many CsIIIs, Cs4000s, and
5071s on the used market tomorrow as you see 4040s and 506Xs today, just as
soon as they trickle down through the same channels as the older units did.

Good point!

So far, of the modern cesium standards, I've only seen a few 5071s on the
used market, and they're still priced pretty high, but I'm certain the
others will trickle down before long.

Most seems to be left-overs from closed labs etc. as far as I have seen.

Eventually, they'll be dirt cheap, because a new CsIII costs half of what a
4040 did when new.  They'll be more accurate, for reasons that you can read
about in 15 years of FCS proceedings, and they'll be more reliable because
of 15 years of electronics evolution and manufacturing improvements.

Good points. Very good points. Thanks for pointing it out. Kind of obvious, but
you need to recall the obvious at times.

Oh, and they'll be more fun, because you'll be able to run Monitor3.  :-)

Indeed. I enjoy doing that. Hmm, which reminds me, I should do something about
setting up a continous logging from my Linux box. Hmm... there is alot of
continous loggings I should be setting up. Time, time... so much to do and so
little time to do it on. Sigh!

Cheers,
Magnus

From: "Robert Lutwak" <RLutwak@Comcast.net> Subject: Re: [time-nuts] Question for the cesium nuts. Date: Fri, 16 Mar 2007 18:09:43 -0400 Message-ID: <001301c76817$cd143aa0$6700a8c0@symmetricom.com> Robert, > There's no reason that you shouldn't see just as many CsIIIs, Cs4000s, and > 5071s on the used market tomorrow as you see 4040s and 506Xs today, just as > soon as they trickle down through the same channels as the older units did. Good point! > So far, of the modern cesium standards, I've only seen a few 5071s on the > used market, and they're still priced pretty high, but I'm certain the > others will trickle down before long. Most seems to be left-overs from closed labs etc. as far as I have seen. > Eventually, they'll be dirt cheap, because a new CsIII costs half of what a > 4040 did when new. They'll be more accurate, for reasons that you can read > about in 15 years of FCS proceedings, and they'll be more reliable because > of 15 years of electronics evolution and manufacturing improvements. Good points. Very good points. Thanks for pointing it out. Kind of obvious, but you need to recall the obvious at times. > Oh, and they'll be more fun, because you'll be able to run Monitor3. :-) Indeed. I enjoy doing that. Hmm, which reminds me, I should do something about setting up a continous logging from my Linux box. Hmm... there is alot of continous loggings I should be setting up. Time, time... so much to do and so little time to do it on. Sigh! Cheers, Magnus
CH
Chuck Harris
Fri, Mar 16, 2007 10:41 PM

Robert Lutwak wrote:

There's no reason that you shouldn't see just as many CsIIIs, Cs4000s, and
5071s on the used market tomorrow as you see 4040s and 506Xs today, just as
soon as they trickle down through the same channels as the older units did.

Well, there is one reason, and that is the US DRMO discovered it is cheaper
to destroy test equipment than it is to sell it.  So, I doubt we will ever
again see the vast supply of surplus electronics that came from the cold war
ramping down.

-Chuck Harris

Robert Lutwak wrote: > There's no reason that you shouldn't see just as many CsIIIs, Cs4000s, and > 5071s on the used market tomorrow as you see 4040s and 506Xs today, just as > soon as they trickle down through the same channels as the older units did. Well, there is one reason, and that is the US DRMO discovered it is cheaper to destroy test equipment than it is to sell it. So, I doubt we will ever again see the vast supply of surplus electronics that came from the cold war ramping down. -Chuck Harris
MW
M. Warner Losh
Fri, Mar 16, 2007 10:46 PM

In message: 45FB1D31.2030200@erols.com
Chuck Harris cfharris@erols.com writes:
: Robert Lutwak wrote:
: > There's no reason that you shouldn't see just as many CsIIIs, Cs4000s, and
: > 5071s on the used market tomorrow as you see 4040s and 506Xs today, just as
: > soon as they trickle down through the same channels as the older units did.
:
: Well, there is one reason, and that is the US DRMO discovered it is cheaper
: to destroy test equipment than it is to sell it.  So, I doubt we will ever
: again see the vast supply of surplus electronics that came from the cold war
: ramping down.

/me wonders how it is cheaper to destory a $25k clock than to sell
it...

Warner

In message: <45FB1D31.2030200@erols.com> Chuck Harris <cfharris@erols.com> writes: : Robert Lutwak wrote: : > There's no reason that you shouldn't see just as many CsIIIs, Cs4000s, and : > 5071s on the used market tomorrow as you see 4040s and 506Xs today, just as : > soon as they trickle down through the same channels as the older units did. : : Well, there is one reason, and that is the US DRMO discovered it is cheaper : to destroy test equipment than it is to sell it. So, I doubt we will ever : again see the vast supply of surplus electronics that came from the cold war : ramping down. /me wonders how it is cheaper to destory a $25k clock than to sell it... Warner
S
shoppa@trailing-edge.com
Sat, Mar 17, 2007 12:38 AM

"Jack Hudler" jack@hudler.org wrote:

Which begs the question of; who's cesium standard will we buy in surplus
market?

From my POV (which could be myopic), a few CBT manufacturers are controlling
what remains of this market (no I'm not a conspiracy nut, it's just
business), so it seems to me that the surplus market is going to get very
thin in the near future.
Supply and demand dictates that surplus market prices will skyrocket out of
the vast majority of amateur reaches in the coming years.

So what's the next cesium standard to start showing up on eBay in numbers
with life left?

The last big boom in CBT and other timing equipment was the telecom
boom of the late 90's, and when it went pop (actually even before) a lot
of almost new stuff hit the surplus stream big-time.

The telecom boom-bust cycle there will not be repeated in quite
the same way and I cannot imagine doing it to the same scale, where
startup telcos and equipment manufacturers could get funding to
get hundreds of millions of dollars equipment made without any
paying customers :-).

The next big thing may not be cesium standards, might be something else,
but it'll probably be tied to some other boom-bust cycle in some
tech/business sector.

Each boom-bust cycle will not be exacltly like the one before, the
cold-war military/avionics stuff I grew up with is probably never
going to happen again either. Nor are the R-390A's piled to the
sky, nor are the command sets by the thousands, but in reality none
of that stuff is particularly hard to come by. I just picked up a
couple of command sets that had been sitting in a guy's garage
for 40 years and are exactly like they were when he got them!

Tim.

"Jack Hudler" <jack@hudler.org> wrote: > Which begs the question of; who's cesium standard will we buy in surplus > market? > > From my POV (which could be myopic), a few CBT manufacturers are controlling > what remains of this market (no I'm not a conspiracy nut, it's just > business), so it seems to me that the surplus market is going to get very > thin in the near future. > Supply and demand dictates that surplus market prices will skyrocket out of > the vast majority of amateur reaches in the coming years. > > So what's the next cesium standard to start showing up on eBay in numbers > with life left? The last big boom in CBT and other timing equipment was the telecom boom of the late 90's, and when it went pop (actually even before) a lot of almost new stuff hit the surplus stream big-time. The telecom boom-bust cycle there will not be repeated in quite the same way and I cannot imagine doing it to the same scale, where startup telcos and equipment manufacturers could get funding to get hundreds of millions of dollars equipment made without any paying customers :-). The next big thing may not be cesium standards, might be something else, but it'll probably be tied to some other boom-bust cycle in some tech/business sector. Each boom-bust cycle will not be exacltly like the one before, the cold-war military/avionics stuff I grew up with is probably never going to happen again either. Nor are the R-390A's piled to the sky, nor are the command sets by the thousands, but in reality none of that stuff is particularly hard to come by. I just picked up a couple of command sets that had been sitting in a guy's garage for 40 years and are exactly like they were when he got them! Tim.