FS
Frank Stellmach
Sat, Oct 26, 2019 8:35 PM
Didier,
if you have the old board with DALLAS RAMs:
- make a dump of the CALRAM into a file, as Poul-Henning explained
- buy the 2kB and 2x 32kB DALLAS RAMs, available from several distributors
- DS1220AD-120+ for about 11$/€ and DS1230Y-150+ for about 22$/€ will
do the job
- unsolder the old RAMs, and solder precision DIL sockets instead
- use a programmer like the TL866 to first make an additional download
from the RAM directly
- program the content of the old CAL RAM to the new 2kB one, and insert
all three new RAMs, maybe add these four 32kB RAMs for additional data
buffer
Alternatively, you can replace the nvSRAMs by FRAMs, description
somewhere in eevblog, or at TiN
if you have the new board with the SnapHat batteries, you can't upload
the RAM content again, but:
- buy 2 SnapHat battery replacements
- switch on the instrument
- remove the old batteries
- insert the new batteries, I would support the PCB with something like
isolated plastics, due to heavy bending forces
video and description at M. J. Lorton (not competent at all, but
instructive)
Frank
Didier,
if you have the old board with DALLAS RAMs:
- make a dump of the CALRAM into a file, as Poul-Henning explained
- buy the 2kB and 2x 32kB DALLAS RAMs, available from several distributors
- DS1220AD-120+ for about 11$/€ and DS1230Y-150+ for about 22$/€ will
do the job
- unsolder the old RAMs, and solder precision DIL sockets instead
- use a programmer like the TL866 to first make an additional download
from the RAM directly
- program the content of the old CAL RAM to the new 2kB one, and insert
all three new RAMs, maybe add these four 32kB RAMs for additional data
buffer
Alternatively, you can replace the nvSRAMs by FRAMs, description
somewhere in eevblog, or at TiN
if you have the new board with the SnapHat batteries, you can't upload
the RAM content again, but:
- buy 2 SnapHat battery replacements
- switch on the instrument
- remove the old batteries
- insert the new batteries, I would support the PCB with something like
isolated plastics, due to heavy bending forces
video and description at M. J. Lorton (not competent at all, but
instructive)
Frank
DJ
Didier Juges
Sat, Oct 26, 2019 10:35 PM
Thanks Frank, it seems like it's time to take the covers off and figure out
what I've got.
On Sat, Oct 26, 2019 at 3:38 PM Frank Stellmach frank.stellmach@freenet.de
wrote:
Didier,
if you have the old board with DALLAS RAMs:
- make a dump of the CALRAM into a file, as Poul-Henning explained
- buy the 2kB and 2x 32kB DALLAS RAMs, available from several distributors
- DS1220AD-120+ for about 11$/€ and DS1230Y-150+ for about 22$/€ will
do the job
- unsolder the old RAMs, and solder precision DIL sockets instead
- use a programmer like the TL866 to first make an additional download
from the RAM directly
- program the content of the old CAL RAM to the new 2kB one, and insert
all three new RAMs, maybe add these four 32kB RAMs for additional data
buffer
Alternatively, you can replace the nvSRAMs by FRAMs, description
somewhere in eevblog, or at TiN
if you have the new board with the SnapHat batteries, you can't upload
the RAM content again, but:
- buy 2 SnapHat battery replacements
- switch on the instrument
- remove the old batteries
- insert the new batteries, I would support the PCB with something like
isolated plastics, due to heavy bending forces
video and description at M. J. Lorton (not competent at all, but
instructive)
Frank
volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.
Thanks Frank, it seems like it's time to take the covers off and figure out
what I've got.
On Sat, Oct 26, 2019 at 3:38 PM Frank Stellmach <frank.stellmach@freenet.de>
wrote:
> Didier,
>
> if you have the old board with DALLAS RAMs:
>
> - make a dump of the CALRAM into a file, as Poul-Henning explained
> - buy the 2kB and 2x 32kB DALLAS RAMs, available from several distributors
> - DS1220AD-120+ for about 11$/€ and DS1230Y-150+ for about 22$/€ will
> do the job
> - unsolder the old RAMs, and solder precision DIL sockets instead
> - use a programmer like the TL866 to first make an additional download
> from the RAM directly
> - program the content of the old CAL RAM to the new 2kB one, and insert
> all three new RAMs, maybe add these four 32kB RAMs for additional data
> buffer
>
> Alternatively, you can replace the nvSRAMs by FRAMs, description
> somewhere in eevblog, or at TiN
>
> if you have the new board with the SnapHat batteries, you can't upload
> the RAM content again, but:
>
> - buy 2 SnapHat battery replacements
> - switch on the instrument
> - remove the old batteries
> - insert the new batteries, I would support the PCB with something like
> isolated plastics, due to heavy bending forces
>
> video and description at M. J. Lorton (not competent at all, but
> instructive)
>
> Frank
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@lists.febo.com
> To unsubscribe, go to
> http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
> and follow the instructions there.
>
DJ
Didier Juges
Sat, Oct 26, 2019 10:45 PM
NVRAMs are Dallas parts with date codes of 2009 and 2010, which explains
the 1934 CALRAM BATTERY sticker on the back of the instrument and may be
the reason it was discarded. The next cal was going to be expensive. Not
socketed.
There are 4 empty sockets next to the NVRAMs, is that for optional
firmware?
Are you saying only the 2kB part contains cal data?
On Sat, Oct 26, 2019 at 5:35 PM Didier Juges shalimr9@gmail.com wrote:
Thanks Frank, it seems like it's time to take the covers off and figure
out what I've got.
On Sat, Oct 26, 2019 at 3:38 PM Frank Stellmach <
frank.stellmach@freenet.de> wrote:
Didier,
if you have the old board with DALLAS RAMs:
- make a dump of the CALRAM into a file, as Poul-Henning explained
- buy the 2kB and 2x 32kB DALLAS RAMs, available from several distributors
- DS1220AD-120+ for about 11$/€ and DS1230Y-150+ for about 22$/€ will
do the job
- unsolder the old RAMs, and solder precision DIL sockets instead
- use a programmer like the TL866 to first make an additional download
from the RAM directly
- program the content of the old CAL RAM to the new 2kB one, and insert
all three new RAMs, maybe add these four 32kB RAMs for additional data
buffer
Alternatively, you can replace the nvSRAMs by FRAMs, description
somewhere in eevblog, or at TiN
if you have the new board with the SnapHat batteries, you can't upload
the RAM content again, but:
- buy 2 SnapHat battery replacements
- switch on the instrument
- remove the old batteries
- insert the new batteries, I would support the PCB with something like
isolated plastics, due to heavy bending forces
video and description at M. J. Lorton (not competent at all, but
instructive)
Frank
volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.
NVRAMs are Dallas parts with date codes of 2009 and 2010, which explains
the 1934 CALRAM BATTERY sticker on the back of the instrument and may be
the reason it was discarded. The next cal was going to be expensive. Not
socketed.
There are 4 empty sockets next to the NVRAMs, is that for optional
firmware?
Are you saying only the 2kB part contains cal data?
On Sat, Oct 26, 2019 at 5:35 PM Didier Juges <shalimr9@gmail.com> wrote:
> Thanks Frank, it seems like it's time to take the covers off and figure
> out what I've got.
>
> On Sat, Oct 26, 2019 at 3:38 PM Frank Stellmach <
> frank.stellmach@freenet.de> wrote:
>
>> Didier,
>>
>> if you have the old board with DALLAS RAMs:
>>
>> - make a dump of the CALRAM into a file, as Poul-Henning explained
>> - buy the 2kB and 2x 32kB DALLAS RAMs, available from several distributors
>> - DS1220AD-120+ for about 11$/€ and DS1230Y-150+ for about 22$/€ will
>> do the job
>> - unsolder the old RAMs, and solder precision DIL sockets instead
>> - use a programmer like the TL866 to first make an additional download
>> from the RAM directly
>> - program the content of the old CAL RAM to the new 2kB one, and insert
>> all three new RAMs, maybe add these four 32kB RAMs for additional data
>> buffer
>>
>> Alternatively, you can replace the nvSRAMs by FRAMs, description
>> somewhere in eevblog, or at TiN
>>
>> if you have the new board with the SnapHat batteries, you can't upload
>> the RAM content again, but:
>>
>> - buy 2 SnapHat battery replacements
>> - switch on the instrument
>> - remove the old batteries
>> - insert the new batteries, I would support the PCB with something like
>> isolated plastics, due to heavy bending forces
>>
>> video and description at M. J. Lorton (not competent at all, but
>> instructive)
>>
>> Frank
>>
>>
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@lists.febo.com
>> To unsubscribe, go to
>> http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
>> and follow the instructions there.
>>
>
IT
Illya Tsemenko
Sat, Oct 26, 2019 11:46 PM
Additional sockets are for extra RAM for option 001. It is useful in digitizing applications if you need it. You can just buy usual SRAMs to add it. No need Dallas NVRAM for extra memory.
Calibration data indeed stored in lonely DS1220. You can backup it for history reasons. I would also recommend replacing electrolytic caps on A4 and A6 PSU boards, and Shaffner mains filter as preventive measure. Latest firmware for 3458 is Rev9, you can find caps BOM and firmware dumps on my site at /fix/hp3458a/ .
After this I strongly recommend to run meter for 3-4 weeks 24/7 and log SN18 service note calibration values , in sequence - store values, run ACAL ALL, store values again. This test can help you to determine if your meter have bad ADC (can be fixed only by replacement, typical issue with 3458s) or other problems with stability, like big DCV tempco. Typical 3458A have tc about 0.2-0.3 ppm/K, but some gems like mine have that below 0.05 ppm/K for 10V range.
If you want to keep meter for metrology purposes, you can opt in to modify A9 LTZ reference for lower oven temperature. That will help you to obtain <2 ppm/year stability at expense of extra noise and invalidated calibration.
BR,
Illya
xDevs.com
On October 26, 2019 6:45:28 PM EDT, Didier Juges shalimr9@gmail.com wrote:
NVRAMs are Dallas parts with date codes of 2009 and 2010, which
explains
the 1934 CALRAM BATTERY sticker on the back of the instrument and may
be
the reason it was discarded. The next cal was going to be expensive.
Not
socketed.
There are 4 empty sockets next to the NVRAMs, is that for optional
firmware?
Are you saying only the 2kB part contains cal data?
On Sat, Oct 26, 2019 at 5:35 PM Didier Juges shalimr9@gmail.com
wrote:
Thanks Frank, it seems like it's time to take the covers off and
Didier,
if you have the old board with DALLAS RAMs:
- make a dump of the CALRAM into a file, as Poul-Henning explained
- buy the 2kB and 2x 32kB DALLAS RAMs, available from several
- DS1220AD-120+ for about 11$/€ and DS1230Y-150+ for about 22$/€
do the job
- unsolder the old RAMs, and solder precision DIL sockets instead
- use a programmer like the TL866 to first make an additional
from the RAM directly
- program the content of the old CAL RAM to the new 2kB one, and
all three new RAMs, maybe add these four 32kB RAMs for additional
buffer
Alternatively, you can replace the nvSRAMs by FRAMs, description
somewhere in eevblog, or at TiN
if you have the new board with the SnapHat batteries, you can't
the RAM content again, but:
- buy 2 SnapHat battery replacements
- switch on the instrument
- remove the old batteries
- insert the new batteries, I would support the PCB with something
Additional sockets are for extra RAM for option 001. It is useful in digitizing applications if you need it. You can just buy usual SRAMs to add it. No need Dallas NVRAM for extra memory.
Calibration data indeed stored in lonely DS1220. You can backup it for history reasons. I would also recommend replacing electrolytic caps on A4 and A6 PSU boards, and Shaffner mains filter as preventive measure. Latest firmware for 3458 is Rev9, you can find caps BOM and firmware dumps on my site at /fix/hp3458a/ .
After this I strongly recommend to run meter for 3-4 weeks 24/7 and log SN18 service note calibration values , in sequence - store values, run ACAL ALL, store values again. This test can help you to determine if your meter have bad ADC (can be fixed only by replacement, typical issue with 3458s) or other problems with stability, like big DCV tempco. Typical 3458A have tc about 0.2-0.3 ppm/K, but some gems like mine have that below 0.05 ppm/K for 10V range.
If you want to keep meter for metrology purposes, you can opt in to modify A9 LTZ reference for lower oven temperature. That will help you to obtain <2 ppm/year stability at expense of extra noise and invalidated calibration.
BR,
Illya
xDevs.com
On October 26, 2019 6:45:28 PM EDT, Didier Juges <shalimr9@gmail.com> wrote:
>NVRAMs are Dallas parts with date codes of 2009 and 2010, which
>explains
>the 1934 CALRAM BATTERY sticker on the back of the instrument and may
>be
>the reason it was discarded. The next cal was going to be expensive.
>Not
>socketed.
>
>There are 4 empty sockets next to the NVRAMs, is that for optional
>firmware?
>
>Are you saying only the 2kB part contains cal data?
>
>
>On Sat, Oct 26, 2019 at 5:35 PM Didier Juges <shalimr9@gmail.com>
>wrote:
>
>> Thanks Frank, it seems like it's time to take the covers off and
>figure
>> out what I've got.
>>
>> On Sat, Oct 26, 2019 at 3:38 PM Frank Stellmach <
>> frank.stellmach@freenet.de> wrote:
>>
>>> Didier,
>>>
>>> if you have the old board with DALLAS RAMs:
>>>
>>> - make a dump of the CALRAM into a file, as Poul-Henning explained
>>> - buy the 2kB and 2x 32kB DALLAS RAMs, available from several
>distributors
>>> - DS1220AD-120+ for about 11$/€ and DS1230Y-150+ for about 22$/€
>will
>>> do the job
>>> - unsolder the old RAMs, and solder precision DIL sockets instead
>>> - use a programmer like the TL866 to first make an additional
>download
>>> from the RAM directly
>>> - program the content of the old CAL RAM to the new 2kB one, and
>insert
>>> all three new RAMs, maybe add these four 32kB RAMs for additional
>data
>>> buffer
>>>
>>> Alternatively, you can replace the nvSRAMs by FRAMs, description
>>> somewhere in eevblog, or at TiN
>>>
>>> if you have the new board with the SnapHat batteries, you can't
>upload
>>> the RAM content again, but:
>>>
>>> - buy 2 SnapHat battery replacements
>>> - switch on the instrument
>>> - remove the old batteries
>>> - insert the new batteries, I would support the PCB with something
>like
>>> isolated plastics, due to heavy bending forces
>>>
>>> video and description at M. J. Lorton (not competent at all, but
>>> instructive)
>>>
>>> Frank
>>>
>>>
>>> _______________________________________________
>>> volt-nuts mailing list -- volt-nuts@lists.febo.com
>>> To unsubscribe, go to
>>> http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
>>> and follow the instructions there.
>>>
>>
>_______________________________________________
>volt-nuts mailing list -- volt-nuts@lists.febo.com
>To unsubscribe, go to
>http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
>and follow the instructions there.
DJ
Didier Juges
Wed, Nov 13, 2019 2:11 AM
Quick progress update:
I was able to read the CALRAM via GPIB using a Prologix adapter and Mark's
software (via KE5FX). Success!
I did not expect the files to be dropped in
C:\Users<login>\AppData\Local\KE5FX... but I found them.
I also got a TL866II+ universal programmer in the mail today. Nice piece of
kit.
I had a Dallas DS1225Y with a date code of 8716 that was on an old board
(yes, I tend to be a packrat) and surprisingly, the part still had all its
data. 32 years later that's pretty good.
A 27C256 EPROM on the same board with a similar date code did not turn out
OK, the TL866 could not read it, yet I know that board was working when it
was put "in storage". How embarrassing the NVRAM outlived the EPROM!
Now the hard part, bite the bullet, open up the 3458 and pull the old parts
off the CPU board. Maybe I'll wait for a long week end, I like to take my
time but I do not like to start something and not finish and have the
equipment in pieces. Somehow these projects rarely turn out good.
In a related subject, I have a couple of Tek TDS754D that could use new
TIMEKEEPER devices, so if the 3458 goes OK, I may do those next.
Didier
On Sat, Oct 26, 2019 at 3:38 PM Frank Stellmach frank.stellmach@freenet.de
wrote:
Didier,
if you have the old board with DALLAS RAMs:
- make a dump of the CALRAM into a file, as Poul-Henning explained
- buy the 2kB and 2x 32kB DALLAS RAMs, available from several distributors
- DS1220AD-120+ for about 11$/€ and DS1230Y-150+ for about 22$/€ will
do the job
- unsolder the old RAMs, and solder precision DIL sockets instead
- use a programmer like the TL866 to first make an additional download
from the RAM directly
- program the content of the old CAL RAM to the new 2kB one, and insert
all three new RAMs, maybe add these four 32kB RAMs for additional data
buffer
Alternatively, you can replace the nvSRAMs by FRAMs, description
somewhere in eevblog, or at TiN
if you have the new board with the SnapHat batteries, you can't upload
the RAM content again, but:
- buy 2 SnapHat battery replacements
- switch on the instrument
- remove the old batteries
- insert the new batteries, I would support the PCB with something like
isolated plastics, due to heavy bending forces
video and description at M. J. Lorton (not competent at all, but
instructive)
Frank
volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.
Quick progress update:
I was able to read the CALRAM via GPIB using a Prologix adapter and Mark's
software (via KE5FX). Success!
I did not expect the files to be dropped in
C:\Users\<login>\AppData\Local\KE5FX\... but I found them.
I also got a TL866II+ universal programmer in the mail today. Nice piece of
kit.
I had a Dallas DS1225Y with a date code of 8716 that was on an old board
(yes, I tend to be a packrat) and surprisingly, the part still had all its
data. 32 years later that's pretty good.
A 27C256 EPROM on the same board with a similar date code did not turn out
OK, the TL866 could not read it, yet I know that board was working when it
was put "in storage". How embarrassing the NVRAM outlived the EPROM!
Now the hard part, bite the bullet, open up the 3458 and pull the old parts
off the CPU board. Maybe I'll wait for a long week end, I like to take my
time but I do not like to start something and not finish and have the
equipment in pieces. Somehow these projects rarely turn out good.
In a related subject, I have a couple of Tek TDS754D that could use new
TIMEKEEPER devices, so if the 3458 goes OK, I may do those next.
Didier
On Sat, Oct 26, 2019 at 3:38 PM Frank Stellmach <frank.stellmach@freenet.de>
wrote:
> Didier,
>
> if you have the old board with DALLAS RAMs:
>
> - make a dump of the CALRAM into a file, as Poul-Henning explained
> - buy the 2kB and 2x 32kB DALLAS RAMs, available from several distributors
> - DS1220AD-120+ for about 11$/€ and DS1230Y-150+ for about 22$/€ will
> do the job
> - unsolder the old RAMs, and solder precision DIL sockets instead
> - use a programmer like the TL866 to first make an additional download
> from the RAM directly
> - program the content of the old CAL RAM to the new 2kB one, and insert
> all three new RAMs, maybe add these four 32kB RAMs for additional data
> buffer
>
> Alternatively, you can replace the nvSRAMs by FRAMs, description
> somewhere in eevblog, or at TiN
>
> if you have the new board with the SnapHat batteries, you can't upload
> the RAM content again, but:
>
> - buy 2 SnapHat battery replacements
> - switch on the instrument
> - remove the old batteries
> - insert the new batteries, I would support the PCB with something like
> isolated plastics, due to heavy bending forces
>
> video and description at M. J. Lorton (not competent at all, but
> instructive)
>
> Frank
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@lists.febo.com
> To unsubscribe, go to
> http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
> and follow the instructions there.
>