time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

PRS-10 input and output calibrations

MD
Magnus Danielson
Sun, Apr 22, 2012 11:49 AM

Fellow time-nuts,

Anyone who has tried to calibrate the PPS input and PPS output
interpolators of the PRS-10?

The output calibration seems pretty easy to do, just a TIC to measure
the PPS to 10 MHz and then sweep through the output delay values to
cover a little more than the 400 ns range.

The input calibration would be something in the similar way, a simpel
method would be to use a coax cable between output and input, and then
use the internal output offset to measure the range of the input
interpolator. However, the input interpolator is a decade better than
the output interpolator. However, considering that I have a 5359A just
waiting to be used, I think I use a fixed output PPS delay and trigger
the 5359A with and then sweep that one.

I would then use the collected data to find suitable correction values.

Comments and suggestions? Experience?

Cheers,
Magnus

Fellow time-nuts, Anyone who has tried to calibrate the PPS input and PPS output interpolators of the PRS-10? The output calibration seems pretty easy to do, just a TIC to measure the PPS to 10 MHz and then sweep through the output delay values to cover a little more than the 400 ns range. The input calibration would be something in the similar way, a simpel method would be to use a coax cable between output and input, and then use the internal output offset to measure the range of the input interpolator. However, the input interpolator is a decade better than the output interpolator. However, considering that I have a 5359A just waiting to be used, I think I use a fixed output PPS delay and trigger the 5359A with and then sweep that one. I would then use the collected data to find suitable correction values. Comments and suggestions? Experience? Cheers, Magnus
B
bg@lysator.liu.se
Sun, Apr 22, 2012 4:12 PM

Magnus,

Interesting! Have a couple of PRS-10s at work that we wanted to lock
together. We never got one to track the other in a reliable way.

--

Björn

Fellow time-nuts,

Anyone who has tried to calibrate the PPS input and PPS output
interpolators of the PRS-10?

The output calibration seems pretty easy to do, just a TIC to measure
the PPS to 10 MHz and then sweep through the output delay values to
cover a little more than the 400 ns range.

The input calibration would be something in the similar way, a simpel
method would be to use a coax cable between output and input, and then
use the internal output offset to measure the range of the input
interpolator. However, the input interpolator is a decade better than
the output interpolator. However, considering that I have a 5359A just
waiting to be used, I think I use a fixed output PPS delay and trigger
the 5359A with and then sweep that one.

I would then use the collected data to find suitable correction values.

Comments and suggestions? Experience?

Cheers,
Magnus


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Magnus, Interesting! Have a couple of PRS-10s at work that we wanted to lock together. We never got one to track the other in a reliable way. -- Björn > Fellow time-nuts, > > Anyone who has tried to calibrate the PPS input and PPS output > interpolators of the PRS-10? > > The output calibration seems pretty easy to do, just a TIC to measure > the PPS to 10 MHz and then sweep through the output delay values to > cover a little more than the 400 ns range. > > The input calibration would be something in the similar way, a simpel > method would be to use a coax cable between output and input, and then > use the internal output offset to measure the range of the input > interpolator. However, the input interpolator is a decade better than > the output interpolator. However, considering that I have a 5359A just > waiting to be used, I think I use a fixed output PPS delay and trigger > the 5359A with and then sweep that one. > > I would then use the collected data to find suitable correction values. > > Comments and suggestions? Experience? > > Cheers, > Magnus > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
MD
Magnus Danielson
Sun, Apr 22, 2012 4:38 PM

Hej Björn,

On 04/22/2012 06:12 PM, bg@lysator.liu.se wrote:

Magnus,

Interesting! Have a couple of PRS-10s at work that we wanted to lock
together. We never got one to track the other in a reliable way.

I'll play around with mine, and when I have more experience you can lend
me them and I can look them over. The PRS-10 is fairly nice since you
can experiment with the soft parameters and only when you are satisfied
you can write them to the EEPROM. You can also read out the EEPROM
parameters regardless of the current running parameters.

The manual isn't particular clever in correlating error observations and
what you should do about it, but it's fairly "open".

Cheers,
Magnus

Hej Björn, On 04/22/2012 06:12 PM, bg@lysator.liu.se wrote: > Magnus, > > Interesting! Have a couple of PRS-10s at work that we wanted to lock > together. We never got one to track the other in a reliable way. I'll play around with mine, and when I have more experience you can lend me them and I can look them over. The PRS-10 is fairly nice since you can experiment with the soft parameters and only when you are satisfied you can write them to the EEPROM. You can also read out the EEPROM parameters regardless of the current running parameters. The manual isn't particular clever in correlating error observations and what you should do about it, but it's fairly "open". Cheers, Magnus
PK
Poul-Henning Kamp
Sun, Apr 22, 2012 6:44 PM

In message 4F93F032.2040608@rubidium.dyndns.org, Magnus Danielson writes:

The input calibration would be something in the similar way,

What I did:

Detune an OCXO slightly (I actually have a 9.99997 MHz OCXO from
IsoTemp), feed it to PPSDIV, disable the discipline code in the
PRS10, collect the measured input time stamps over some hours.

Either you get a nice ramp, or you get som kind of demented staircase,
in which case you try to figure out which of the calibration constants
to mess with.

An alternative is to feed the PRS10 output to a HP333x Synthesizer
and have that generate your 10-epsilon MHz for the PPSdiv.

This general "vernier" method can be used to measure all sorts of tricky
stuff, from interrupt latencies in operating system kernels to
stuff like the above.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

In message <4F93F032.2040608@rubidium.dyndns.org>, Magnus Danielson writes: >The input calibration would be something in the similar way, What I did: Detune an OCXO slightly (I actually have a 9.99997 MHz OCXO from IsoTemp), feed it to PPSDIV, disable the discipline code in the PRS10, collect the measured input time stamps over some hours. Either you get a nice ramp, or you get som kind of demented staircase, in which case you try to figure out which of the calibration constants to mess with. An alternative is to feed the PRS10 output to a HP333x Synthesizer and have that generate your 10-epsilon MHz for the PPSdiv. This general "vernier" method can be used to measure all sorts of tricky stuff, from interrupt latencies in operating system kernels to stuff like the above. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
AB
Azelio Boriani
Sun, Apr 22, 2012 8:04 PM

Björn, what do you mean with "We never got one to track the other in a
reliable way"?
Thai is, how can it be that a PRS10 cannot track another PRS10? What do you
get when trying to track one with the other?

On Sun, Apr 22, 2012 at 8:44 PM, Poul-Henning Kamp phk@phk.freebsd.dkwrote:

In message 4F93F032.2040608@rubidium.dyndns.org, Magnus Danielson
writes:

The input calibration would be something in the similar way,

What I did:

Detune an OCXO slightly (I actually have a 9.99997 MHz OCXO from
IsoTemp), feed it to PPSDIV, disable the discipline code in the
PRS10, collect the measured input time stamps over some hours.

Either you get a nice ramp, or you get som kind of demented staircase,
in which case you try to figure out which of the calibration constants
to mess with.

An alternative is to feed the PRS10 output to a HP333x Synthesizer
and have that generate your 10-epsilon MHz for the PPSdiv.

This general "vernier" method can be used to measure all sorts of tricky
stuff, from interrupt latencies in operating system kernels to
stuff like the above.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Björn, what do you mean with "We never got one to track the other in a reliable way"? Thai is, how can it be that a PRS10 cannot track another PRS10? What do you get when trying to track one with the other? On Sun, Apr 22, 2012 at 8:44 PM, Poul-Henning Kamp <phk@phk.freebsd.dk>wrote: > In message <4F93F032.2040608@rubidium.dyndns.org>, Magnus Danielson > writes: > > >The input calibration would be something in the similar way, > > What I did: > > Detune an OCXO slightly (I actually have a 9.99997 MHz OCXO from > IsoTemp), feed it to PPSDIV, disable the discipline code in the > PRS10, collect the measured input time stamps over some hours. > > Either you get a nice ramp, or you get som kind of demented staircase, > in which case you try to figure out which of the calibration constants > to mess with. > > An alternative is to feed the PRS10 output to a HP333x Synthesizer > and have that generate your 10-epsilon MHz for the PPSdiv. > > This general "vernier" method can be used to measure all sorts of tricky > stuff, from interrupt latencies in operating system kernels to > stuff like the above. > > -- > Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 > phk@FreeBSD.ORG | TCP/IP since RFC 956 > FreeBSD committer | BSD since 4.3-tahoe > Never attribute to malice what can adequately be explained by incompetence. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
B
bg@lysator.liu.se
Sun, Apr 22, 2012 8:47 PM

With factory default settings with 2 PRS-10s, connecting 1PPS_out from one
unit to 1PPS_in on the other, would not align the 1PPS_out pulses. They
were off by several hundred of ns. It was probably an operator error
somewhere. We just did not find the error in the time frame we had
available.

--

Björn

Björn, what do you mean with "We never got one to track the other in a
reliable way"?
Thai is, how can it be that a PRS10 cannot track another PRS10? What do
you
get when trying to track one with the other?

On Sun, Apr 22, 2012 at 8:44 PM, Poul-Henning Kamp
phk@phk.freebsd.dkwrote:

In message 4F93F032.2040608@rubidium.dyndns.org, Magnus Danielson
writes:

The input calibration would be something in the similar way,

What I did:

Detune an OCXO slightly (I actually have a 9.99997 MHz OCXO from
IsoTemp), feed it to PPSDIV, disable the discipline code in the
PRS10, collect the measured input time stamps over some hours.

Either you get a nice ramp, or you get som kind of demented staircase,
in which case you try to figure out which of the calibration constants
to mess with.

An alternative is to feed the PRS10 output to a HP333x Synthesizer
and have that generate your 10-epsilon MHz for the PPSdiv.

This general "vernier" method can be used to measure all sorts of tricky
stuff, from interrupt latencies in operating system kernels to
stuff like the above.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by
incompetence.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

With factory default settings with 2 PRS-10s, connecting 1PPS_out from one unit to 1PPS_in on the other, would not align the 1PPS_out pulses. They were off by several hundred of ns. It was probably an operator error somewhere. We just did not find the error in the time frame we had available. -- Björn > Björn, what do you mean with "We never got one to track the other in a > reliable way"? > Thai is, how can it be that a PRS10 cannot track another PRS10? What do > you > get when trying to track one with the other? > > On Sun, Apr 22, 2012 at 8:44 PM, Poul-Henning Kamp > <phk@phk.freebsd.dk>wrote: > >> In message <4F93F032.2040608@rubidium.dyndns.org>, Magnus Danielson >> writes: >> >> >The input calibration would be something in the similar way, >> >> What I did: >> >> Detune an OCXO slightly (I actually have a 9.99997 MHz OCXO from >> IsoTemp), feed it to PPSDIV, disable the discipline code in the >> PRS10, collect the measured input time stamps over some hours. >> >> Either you get a nice ramp, or you get som kind of demented staircase, >> in which case you try to figure out which of the calibration constants >> to mess with. >> >> An alternative is to feed the PRS10 output to a HP333x Synthesizer >> and have that generate your 10-epsilon MHz for the PPSdiv. >> >> This general "vernier" method can be used to measure all sorts of tricky >> stuff, from interrupt latencies in operating system kernels to >> stuff like the above. >> >> -- >> Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 >> phk@FreeBSD.ORG | TCP/IP since RFC 956 >> FreeBSD committer | BSD since 4.3-tahoe >> Never attribute to malice what can adequately be explained by >> incompetence. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
AB
Azelio Boriani
Sun, Apr 22, 2012 9:40 PM

That is very interesting in my opinion and, for those who have access to
more than one PRS10, further investigation should be done. I had a new
PRS10 and, just out-of-the-box, it locked (not instantly, of course) to the
HP53508 PPS.

On Sun, Apr 22, 2012 at 10:47 PM, bg@lysator.liu.se wrote:

With factory default settings with 2 PRS-10s, connecting 1PPS_out from one
unit to 1PPS_in on the other, would not align the 1PPS_out pulses. They
were off by several hundred of ns. It was probably an operator error
somewhere. We just did not find the error in the time frame we had
available.

--

Björn

Björn, what do you mean with "We never got one to track the other in a
reliable way"?
Thai is, how can it be that a PRS10 cannot track another PRS10? What do
you
get when trying to track one with the other?

On Sun, Apr 22, 2012 at 8:44 PM, Poul-Henning Kamp
phk@phk.freebsd.dkwrote:

In message 4F93F032.2040608@rubidium.dyndns.org, Magnus Danielson
writes:

The input calibration would be something in the similar way,

What I did:

Detune an OCXO slightly (I actually have a 9.99997 MHz OCXO from
IsoTemp), feed it to PPSDIV, disable the discipline code in the
PRS10, collect the measured input time stamps over some hours.

Either you get a nice ramp, or you get som kind of demented staircase,
in which case you try to figure out which of the calibration constants
to mess with.

An alternative is to feed the PRS10 output to a HP333x Synthesizer
and have that generate your 10-epsilon MHz for the PPSdiv.

This general "vernier" method can be used to measure all sorts of tricky
stuff, from interrupt latencies in operating system kernels to
stuff like the above.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by
incompetence.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

That is very interesting in my opinion and, for those who have access to more than one PRS10, further investigation should be done. I had a new PRS10 and, just out-of-the-box, it locked (not instantly, of course) to the HP53508 PPS. On Sun, Apr 22, 2012 at 10:47 PM, <bg@lysator.liu.se> wrote: > With factory default settings with 2 PRS-10s, connecting 1PPS_out from one > unit to 1PPS_in on the other, would not align the 1PPS_out pulses. They > were off by several hundred of ns. It was probably an operator error > somewhere. We just did not find the error in the time frame we had > available. > > -- > > Björn > > > > Björn, what do you mean with "We never got one to track the other in a > > reliable way"? > > Thai is, how can it be that a PRS10 cannot track another PRS10? What do > > you > > get when trying to track one with the other? > > > > On Sun, Apr 22, 2012 at 8:44 PM, Poul-Henning Kamp > > <phk@phk.freebsd.dk>wrote: > > > >> In message <4F93F032.2040608@rubidium.dyndns.org>, Magnus Danielson > >> writes: > >> > >> >The input calibration would be something in the similar way, > >> > >> What I did: > >> > >> Detune an OCXO slightly (I actually have a 9.99997 MHz OCXO from > >> IsoTemp), feed it to PPSDIV, disable the discipline code in the > >> PRS10, collect the measured input time stamps over some hours. > >> > >> Either you get a nice ramp, or you get som kind of demented staircase, > >> in which case you try to figure out which of the calibration constants > >> to mess with. > >> > >> An alternative is to feed the PRS10 output to a HP333x Synthesizer > >> and have that generate your 10-epsilon MHz for the PPSdiv. > >> > >> This general "vernier" method can be used to measure all sorts of tricky > >> stuff, from interrupt latencies in operating system kernels to > >> stuff like the above. > >> > >> -- > >> Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 > >> phk@FreeBSD.ORG | TCP/IP since RFC 956 > >> FreeBSD committer | BSD since 4.3-tahoe > >> Never attribute to malice what can adequately be explained by > >> incompetence. > >> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@febo.com > >> To unsubscribe, go to > >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> and follow the instructions there. > >> > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
MD
Magnus Danielson
Sun, Apr 22, 2012 10:02 PM

On 04/22/2012 10:47 PM, bg@lysator.liu.se wrote:

With factory default settings with 2 PRS-10s, connecting 1PPS_out from one
unit to 1PPS_in on the other, would not align the 1PPS_out pulses. They
were off by several hundred of ns. It was probably an operator error
somewhere. We just did not find the error in the time frame we had
available.

Interlocking two oscillators like that might be a really bad idea, as
the PLL will have a gain, so risk having a gain loop, especially when
the oscillators have the same default parameters being setup. This is
similar to the form of long PLL chain analysis that has been simulated
using a single PLL and a delay-loop. The peaking gain at the PLL
bandwidth is what kills it.

If there is input and/or output non-linearities in the latest
calibration parameters, that may be a reason they locked up in a strange
situation.

Using a common PPS pulse as input and then compare the output PPS pulses
is a much more valid test.

Cheers,
Magnus

On 04/22/2012 10:47 PM, bg@lysator.liu.se wrote: > With factory default settings with 2 PRS-10s, connecting 1PPS_out from one > unit to 1PPS_in on the other, would not align the 1PPS_out pulses. They > were off by several hundred of ns. It was probably an operator error > somewhere. We just did not find the error in the time frame we had > available. Interlocking two oscillators like that might be a really bad idea, as the PLL will have a gain, so risk having a gain loop, especially when the oscillators have the same default parameters being setup. This is similar to the form of long PLL chain analysis that has been simulated using a single PLL and a delay-loop. The peaking gain at the PLL bandwidth is what kills it. If there is input and/or output non-linearities in the latest calibration parameters, that may be a reason they locked up in a strange situation. Using a common PPS pulse as input and then compare the output PPS pulses is a much more valid test. Cheers, Magnus
PK
Poul-Henning Kamp
Mon, Apr 23, 2012 6:15 AM

That is very interesting in my opinion and, for those who have access to
more than one PRS10, further investigation should be done. I had a new
PRS10 and, just out-of-the-box, it locked (not instantly, of course) to the
HP53508 PPS.

There are threel obvious issues:  Which flank you trigger on and
which polarity and width the output pulse has.

The output pulse from the PRS10 is pretty narrow, making it very
easy to mistakenly use the wrong flank.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

In message <CAL8XPmO0Wj2AUW8Q=-DrDGGDMdhTsteVJaMKgkXktgq=0J4CSA@mail.gmail.com> , Azelio Boriani writes: >That is very interesting in my opinion and, for those who have access to >more than one PRS10, further investigation should be done. I had a new >PRS10 and, just out-of-the-box, it locked (not instantly, of course) to the >HP53508 PPS. There are threel obvious issues: Which flank you trigger on and which polarity and width the output pulse has. The output pulse from the PRS10 is pretty narrow, making it very easy to mistakenly use the wrong flank. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
PK
Poul-Henning Kamp
Mon, Apr 23, 2012 6:16 AM

In message 4F947FF0.8070703@rubidium.dyndns.org, Magnus Danielson writes:

On 04/22/2012 10:47 PM, bg@lysator.liu.se wrote:

Interlocking two oscillators like that might be a really bad idea, as
the PLL will have a gain, so risk having a gain loop, especially when
the oscillators have the same default parameters being setup.

You would clearly want to torque the timeconstant of the second PRS10
in that scenario, but idea is not without merit.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

In message <4F947FF0.8070703@rubidium.dyndns.org>, Magnus Danielson writes: >On 04/22/2012 10:47 PM, bg@lysator.liu.se wrote: >Interlocking two oscillators like that might be a really bad idea, as >the PLL will have a gain, so risk having a gain loop, especially when >the oscillators have the same default parameters being setup. You would clearly want to torque the timeconstant of the second PRS10 in that scenario, but idea is not without merit. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
MD
Magnus Danielson
Mon, Apr 23, 2012 8:24 PM

On 04/23/2012 08:16 AM, Poul-Henning Kamp wrote:

In message4F947FF0.8070703@rubidium.dyndns.org, Magnus Danielson writes:

On 04/22/2012 10:47 PM, bg@lysator.liu.se wrote:

Interlocking two oscillators like that might be a really bad idea, as
the PLL will have a gain, so risk having a gain loop, especially when
the oscillators have the same default parameters being setup.

You would clearly want to torque the timeconstant of the second PRS10
in that scenario, but idea is not without merit.

It helps a bit. You can build phase margin that way if not the gain
margin is on your side.

Cheers,
Magnus

On 04/23/2012 08:16 AM, Poul-Henning Kamp wrote: > In message<4F947FF0.8070703@rubidium.dyndns.org>, Magnus Danielson writes: >> On 04/22/2012 10:47 PM, bg@lysator.liu.se wrote: > >> Interlocking two oscillators like that might be a really bad idea, as >> the PLL will have a gain, so risk having a gain loop, especially when >> the oscillators have the same default parameters being setup. > > You would clearly want to torque the timeconstant of the second PRS10 > in that scenario, but idea is not without merit. It helps a bit. You can build phase margin that way if not the gain margin is on your side. Cheers, Magnus