5061A HV Supply, et al.

PS
paul swed
Mon, Nov 14, 2022 8:25 PM

Jim
I may be missing something here. Your suggesting the 3500 V ion pump supply
is bad.
But I believe the ion pump meter is pegged from an earlier post. If true
then the 3500 V is fine. If the tubes dirty and drawing too much current
there is a lockout circuit on a15 that will not allow the CS oven and 2500
V to come up until the ion pump is below about 15-20ua on the meter. It
senses the current being drawn by the 3500 HV. Enters a15 on pin 1 and
inhibits the rest of the system.
If my thoughts are correct then you need to pump down the tube as Rodger
says with a 3500 V supply and as I suggested 300 ua is way more than
enough. There is a chance the tube may work in a relative quality sense.
That is it could lock but there are so few Cs that the output of the system
is noisy.

Regards
Paul
WB8TSL

On Mon, Nov 14, 2022 at 3:00 PM Jim Muehlberg via time-nuts <
time-nuts@lists.febo.com> wrote:

Rodger,

Thanks for the encouragement!  I'm trying not to spend too much effort
on this - sort of a distraction from real work.  One would think that
NRAO would have a HV power supply lying around somewhere, but astronomy
has gone solid state now for decades!  We have a grad student trying to
nurse along another 5061 and we'd like to have a pair. I guess I'll
slice open the A18 module and see what makes it tick.

Thanks again,

Jim

On 11/10/2022 9:11 AM, Rodger via time-nuts wrote:

Hey Jim,

I got a 5061a a couple years ago with high ION pump current.  In my case

the 3500 volts was there but the tube was not pumping down, even after
being on for weeks.  I found a Terranova 741 Ion pump power supply on eBay
for under $100.  It was listed as not working but worked AOK when I got
it.  It's a nice supply with variable output voltage, current limiting,
safety interlocks, etc.  I hooked it up to the tube and let it run for a
few weeks.  Tube pumped down successfully and is working AOK now.  My guess
is that my A18 wasn't capable of supplying the current needed to pump down
the tube but the external power supply was.

Note: I proceeded very carefully with the external supply, letting it

run for days at lower voltages and gradually working up to 3500.  Also, I
kept the current limit set as low as it would go.  Anyway, it worked for
me.  Good luck with yours.

Rodger

-----Original Message-----
From: Jim Muehlberg via time-nuts time-nuts@lists.febo.com
Sent: Wednesday, November 9, 2022 11:07 AM
To: time-nuts@lists.febo.com
Cc: Jim Muehlberg jmuehlbe@nrao.edu
Subject: [time-nuts] 5061A HV Supply, et al.

I've recently come into possession of a 5061A.  It is claimed that the

tube is exhausted, but I'm unwilling to scrap it (horrors!) yet.  I'm not
well equipped for HV measurements, but it appears the +3500 V supply is not
working.  So, I'm hoping the tube just needs to be pumped since it hasn't
been powered for years.  Also missing the A3 assembly, but I'll worry about
that later...

Anyone willing to part with their A18 or loan it to me?

Thanks!

--

Jim Muehlberg
Senior Engineer

National Radio Astronomy Observatory
ngVLA Local Oscillator Lead

1180 Boxwood Estates Rd B-111
Charlottesville, VA 22903-4602
P 434.296.0270
F 434.296.0324
www.cv.nrao.edu/~jmuehlbe


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

Jim I may be missing something here. Your suggesting the 3500 V ion pump supply is bad. But I believe the ion pump meter is pegged from an earlier post. If true then the 3500 V is fine. If the tubes dirty and drawing too much current there is a lockout circuit on a15 that will not allow the CS oven and 2500 V to come up until the ion pump is below about 15-20ua on the meter. It senses the current being drawn by the 3500 HV. Enters a15 on pin 1 and inhibits the rest of the system. If my thoughts are correct then you need to pump down the tube as Rodger says with a 3500 V supply and as I suggested 300 ua is way more than enough. There is a chance the tube may work in a relative quality sense. That is it could lock but there are so few Cs that the output of the system is noisy. Regards Paul WB8TSL On Mon, Nov 14, 2022 at 3:00 PM Jim Muehlberg via time-nuts < time-nuts@lists.febo.com> wrote: > Rodger, > > Thanks for the encouragement! I'm trying not to spend too much effort > on this - sort of a distraction from real work. One would think that > NRAO would have a HV power supply lying around somewhere, but astronomy > has gone solid state now for decades! We have a grad student trying to > nurse along another 5061 and we'd like to have a pair. I guess I'll > slice open the A18 module and see what makes it tick. > > Thanks again, > > Jim > > On 11/10/2022 9:11 AM, Rodger via time-nuts wrote: > > Hey Jim, > > > > I got a 5061a a couple years ago with high ION pump current. In my case > the 3500 volts was there but the tube was not pumping down, even after > being on for weeks. I found a Terranova 741 Ion pump power supply on eBay > for under $100. It was listed as not working but worked AOK when I got > it. It's a nice supply with variable output voltage, current limiting, > safety interlocks, etc. I hooked it up to the tube and let it run for a > few weeks. Tube pumped down successfully and is working AOK now. My guess > is that my A18 wasn't capable of supplying the current needed to pump down > the tube but the external power supply was. > > > > Note: I proceeded very carefully with the external supply, letting it > run for days at lower voltages and gradually working up to 3500. Also, I > kept the current limit set as low as it would go. Anyway, it worked for > me. Good luck with yours. > > > > Rodger > > > > -----Original Message----- > > From: Jim Muehlberg via time-nuts <time-nuts@lists.febo.com> > > Sent: Wednesday, November 9, 2022 11:07 AM > > To: time-nuts@lists.febo.com > > Cc: Jim Muehlberg <jmuehlbe@nrao.edu> > > Subject: [time-nuts] 5061A HV Supply, et al. > > > > I've recently come into possession of a 5061A. It is claimed that the > tube is exhausted, but I'm unwilling to scrap it (horrors!) yet. I'm not > well equipped for HV measurements, but it appears the +3500 V supply is not > working. So, I'm hoping the tube just needs to be pumped since it hasn't > been powered for years. Also missing the A3 assembly, but I'll worry about > that later... > > > > Anyone willing to part with their A18 or loan it to me? > > > > Thanks! > > > -- > > Jim Muehlberg > Senior Engineer > > National Radio Astronomy Observatory > ngVLA Local Oscillator Lead > > 1180 Boxwood Estates Rd B-111 > Charlottesville, VA 22903-4602 > P 434.296.0270 > F 434.296.0324 > www.cv.nrao.edu/~jmuehlbe > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com
LV
Lester Veenstra
Mon, Nov 14, 2022 8:40 PM

Sacrifice an old microwave oven to the Time Gods for its transformer

Lester B Veenstra  K1YCM  MØYCM  W8YCM  6Y6Y W8YCM/6Y 6Y8LV (Reformed USNSG
CTM1)
lester@veenstras.com

452 Stable Ln
Keyser WV 26726 USA

GPS: 39.336826 N  78.982287 W (Google)
GPS: 39.33682 N  78.9823741 W (GPSDO)

Telephones:
Home:            +1-304-289-6057
US cell          +1-304-790-9192
Jamaica cell:   +1-876-456-8898

-----Original Message-----
From: Andy Talbot via time-nuts [mailto:time-nuts@lists.febo.com]
Sent: Monday, November 14, 2022 3:06 PM
To: Discussion of precise time and frequency measurement
Cc: Jim Muehlberg; Andy Talbot
Subject: [time-nuts] Re: 5061A HV Supply, et al. {External}

A simple flyback SMPSU with a Cockcroft-Walton tied on the end will give
you a simple high voltage supply.  A chain of high value resistors and you
can even stabilise it.

Andy
www.g4jnt.com

On Mon, 14 Nov 2022 at 20:00, Jim Muehlberg via time-nuts <
time-nuts@lists.febo.com> wrote:

Rodger,

Thanks for the encouragement!  I'm trying not to spend too much effort
on this - sort of a distraction from real work.  One would think that
NRAO would have a HV power supply lying around somewhere, but astronomy
has gone solid state now for decades!  We have a grad student trying to
nurse along another 5061 and we'd like to have a pair. I guess I'll
slice open the A18 module and see what makes it tick.

Thanks again,

Jim

On 11/10/2022 9:11 AM, Rodger via time-nuts wrote:

Hey Jim,

I got a 5061a a couple years ago with high ION pump current.  In my case

the 3500 volts was there but the tube was not pumping down, even after
being on for weeks.  I found a Terranova 741 Ion pump power supply on eBay
for under $100.  It was listed as not working but worked AOK when I got
it.  It's a nice supply with variable output voltage, current limiting,
safety interlocks, etc.  I hooked it up to the tube and let it run for a
few weeks.  Tube pumped down successfully and is working AOK now.  My

guess

is that my A18 wasn't capable of supplying the current needed to pump down
the tube but the external power supply was.

Note: I proceeded very carefully with the external supply, letting it

run for days at lower voltages and gradually working up to 3500.  Also, I
kept the current limit set as low as it would go.  Anyway, it worked for
me.  Good luck with yours.

Rodger

-----Original Message-----
From: Jim Muehlberg via time-nuts time-nuts@lists.febo.com
Sent: Wednesday, November 9, 2022 11:07 AM
To: time-nuts@lists.febo.com
Cc: Jim Muehlberg jmuehlbe@nrao.edu
Subject: [time-nuts] 5061A HV Supply, et al.

I've recently come into possession of a 5061A.  It is claimed that the

tube is exhausted, but I'm unwilling to scrap it (horrors!) yet.  I'm not
well equipped for HV measurements, but it appears the +3500 V supply is

not

working.  So, I'm hoping the tube just needs to be pumped since it hasn't
been powered for years.  Also missing the A3 assembly, but I'll worry

about

that later...

Anyone willing to part with their A18 or loan it to me?

Thanks!

--

Jim Muehlberg
Senior Engineer

National Radio Astronomy Observatory
ngVLA Local Oscillator Lead

1180 Boxwood Estates Rd B-111
Charlottesville, VA 22903-4602
P 434.296.0270
F 434.296.0324
www.cv.nrao.edu/~jmuehlbe


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

Sacrifice an old microwave oven to the Time Gods for its transformer Lester B Veenstra  K1YCM MØYCM W8YCM 6Y6Y W8YCM/6Y 6Y8LV (Reformed USNSG CTM1) lester@veenstras.com 452 Stable Ln Keyser WV 26726 USA GPS: 39.336826 N  78.982287 W (Google) GPS: 39.33682 N  78.9823741 W (GPSDO) Telephones: Home:            +1-304-289-6057 US cell          +1-304-790-9192 Jamaica cell:  +1-876-456-8898 -----Original Message----- From: Andy Talbot via time-nuts [mailto:time-nuts@lists.febo.com] Sent: Monday, November 14, 2022 3:06 PM To: Discussion of precise time and frequency measurement Cc: Jim Muehlberg; Andy Talbot Subject: [time-nuts] Re: 5061A HV Supply, et al. {External} A simple flyback SMPSU with a Cockcroft-Walton tied on the end will give you a simple high voltage supply. A chain of high value resistors and you can even stabilise it. Andy www.g4jnt.com On Mon, 14 Nov 2022 at 20:00, Jim Muehlberg via time-nuts < time-nuts@lists.febo.com> wrote: > Rodger, > > Thanks for the encouragement! I'm trying not to spend too much effort > on this - sort of a distraction from real work. One would think that > NRAO would have a HV power supply lying around somewhere, but astronomy > has gone solid state now for decades! We have a grad student trying to > nurse along another 5061 and we'd like to have a pair. I guess I'll > slice open the A18 module and see what makes it tick. > > Thanks again, > > Jim > > On 11/10/2022 9:11 AM, Rodger via time-nuts wrote: > > Hey Jim, > > > > I got a 5061a a couple years ago with high ION pump current. In my case > the 3500 volts was there but the tube was not pumping down, even after > being on for weeks. I found a Terranova 741 Ion pump power supply on eBay > for under $100. It was listed as not working but worked AOK when I got > it. It's a nice supply with variable output voltage, current limiting, > safety interlocks, etc. I hooked it up to the tube and let it run for a > few weeks. Tube pumped down successfully and is working AOK now. My guess > is that my A18 wasn't capable of supplying the current needed to pump down > the tube but the external power supply was. > > > > Note: I proceeded very carefully with the external supply, letting it > run for days at lower voltages and gradually working up to 3500. Also, I > kept the current limit set as low as it would go. Anyway, it worked for > me. Good luck with yours. > > > > Rodger > > > > -----Original Message----- > > From: Jim Muehlberg via time-nuts <time-nuts@lists.febo.com> > > Sent: Wednesday, November 9, 2022 11:07 AM > > To: time-nuts@lists.febo.com > > Cc: Jim Muehlberg <jmuehlbe@nrao.edu> > > Subject: [time-nuts] 5061A HV Supply, et al. > > > > I've recently come into possession of a 5061A. It is claimed that the > tube is exhausted, but I'm unwilling to scrap it (horrors!) yet. I'm not > well equipped for HV measurements, but it appears the +3500 V supply is not > working. So, I'm hoping the tube just needs to be pumped since it hasn't > been powered for years. Also missing the A3 assembly, but I'll worry about > that later... > > > > Anyone willing to part with their A18 or loan it to me? > > > > Thanks! > > > -- > > Jim Muehlberg > Senior Engineer > > National Radio Astronomy Observatory > ngVLA Local Oscillator Lead > > 1180 Boxwood Estates Rd B-111 > Charlottesville, VA 22903-4602 > P 434.296.0270 > F 434.296.0324 > www.cv.nrao.edu/~jmuehlbe > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com _______________________________________________ time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe send an email to time-nuts-leave@lists.febo.com
AT
Andy Talbot
Mon, Nov 14, 2022 8:44 PM

I've got three transformers going spare ... :-)

Andy  G4JNT
http://www.g4jnt.comIO90IV58AK12 (and others)

On Mon, 14 Nov 2022 at 20:40, Lester Veenstra m0ycm@veenstras.com wrote:

Sacrifice an old microwave oven to the Time Gods for its transformer

GPS: 39.336826 N  78.982287 W (Google)  =                    FM09MI20DU01
GPS: 39.33682 N  78.9823741 W (GPSDO) =                    FM09MI20CU70

Telephones:
Home:            +1-304-289-6057
US cell          +1-304-790-9192
Jamaica cell:    +1-876-456-8898

-----Original Message-----
From: Andy Talbot via time-nuts [mailto:time-nuts@lists.febo.com]
Sent: Monday, November 14, 2022 3:06 PM
To: Discussion of precise time and frequency measurement
Cc: Jim Muehlberg; Andy Talbot
Subject: [time-nuts] Re: 5061A HV Supply, et al. {External}

A simple flyback SMPSU with a Cockcroft-Walton tied on the end will give
you a simple high voltage supply.  A chain of high value resistors and you
can even stabilise it.

Andy
www.g4jnt.com

On Mon, 14 Nov 2022 at 20:00, Jim Muehlberg via time-nuts <
time-nuts@lists.febo.com> wrote:

Rodger,

Thanks for the encouragement!  I'm trying not to spend too much effort
on this - sort of a distraction from real work.  One would think that
NRAO would have a HV power supply lying around somewhere, but astronomy
has gone solid state now for decades!  We have a grad student trying to
nurse along another 5061 and we'd like to have a pair. I guess I'll
slice open the A18 module and see what makes it tick.

Thanks again,

Jim

On 11/10/2022 9:11 AM, Rodger via time-nuts wrote:

Hey Jim,

I got a 5061a a couple years ago with high ION pump current.  In my

case

the 3500 volts was there but the tube was not pumping down, even after
being on for weeks.  I found a Terranova 741 Ion pump power supply on

eBay

for under $100.  It was listed as not working but worked AOK when I got
it.  It's a nice supply with variable output voltage, current limiting,
safety interlocks, etc.  I hooked it up to the tube and let it run for a
few weeks.  Tube pumped down successfully and is working AOK now.  My

guess

is that my A18 wasn't capable of supplying the current needed to pump

down

the tube but the external power supply was.

Note: I proceeded very carefully with the external supply, letting it

run for days at lower voltages and gradually working up to 3500.  Also, I
kept the current limit set as low as it would go.  Anyway, it worked for
me.  Good luck with yours.

Rodger

-----Original Message-----
From: Jim Muehlberg via time-nuts time-nuts@lists.febo.com
Sent: Wednesday, November 9, 2022 11:07 AM
To: time-nuts@lists.febo.com
Cc: Jim Muehlberg jmuehlbe@nrao.edu
Subject: [time-nuts] 5061A HV Supply, et al.

I've recently come into possession of a 5061A.  It is claimed that the

tube is exhausted, but I'm unwilling to scrap it (horrors!) yet.  I'm not
well equipped for HV measurements, but it appears the +3500 V supply is

not

working.  So, I'm hoping the tube just needs to be pumped since it hasn't
been powered for years.  Also missing the A3 assembly, but I'll worry

about

that later...

Anyone willing to part with their A18 or loan it to me?

Thanks!

--

Jim Muehlberg
Senior Engineer

National Radio Astronomy Observatory
ngVLA Local Oscillator Lead

1180 Boxwood Estates Rd B-111
Charlottesville, VA 22903-4602
P 434.296.0270
F 434.296.0324
www.cv.nrao.edu/~jmuehlbe


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

I've got three transformers going spare ... :-) Andy G4JNT <http://www.g4jnt.com>IO90IV58AK12 (and others) On Mon, 14 Nov 2022 at 20:40, Lester Veenstra <m0ycm@veenstras.com> wrote: > Sacrifice an old microwave oven to the Time Gods for its transformer > > GPS: 39.336826 N 78.982287 W (Google) = FM09MI20DU01 > GPS: 39.33682 N 78.9823741 W (GPSDO) = FM09MI20CU70 > > > > Telephones: > Home: +1-304-289-6057 > US cell +1-304-790-9192 > Jamaica cell: +1-876-456-8898 > > -----Original Message----- > From: Andy Talbot via time-nuts [mailto:time-nuts@lists.febo.com] > Sent: Monday, November 14, 2022 3:06 PM > To: Discussion of precise time and frequency measurement > Cc: Jim Muehlberg; Andy Talbot > Subject: [time-nuts] Re: 5061A HV Supply, et al. {External} > > A simple flyback SMPSU with a Cockcroft-Walton tied on the end will give > you a simple high voltage supply. A chain of high value resistors and you > can even stabilise it. > > Andy > www.g4jnt.com > > > > On Mon, 14 Nov 2022 at 20:00, Jim Muehlberg via time-nuts < > time-nuts@lists.febo.com> wrote: > > > Rodger, > > > > Thanks for the encouragement! I'm trying not to spend too much effort > > on this - sort of a distraction from real work. One would think that > > NRAO would have a HV power supply lying around somewhere, but astronomy > > has gone solid state now for decades! We have a grad student trying to > > nurse along another 5061 and we'd like to have a pair. I guess I'll > > slice open the A18 module and see what makes it tick. > > > > Thanks again, > > > > Jim > > > > On 11/10/2022 9:11 AM, Rodger via time-nuts wrote: > > > Hey Jim, > > > > > > I got a 5061a a couple years ago with high ION pump current. In my > case > > the 3500 volts was there but the tube was not pumping down, even after > > being on for weeks. I found a Terranova 741 Ion pump power supply on > eBay > > for under $100. It was listed as not working but worked AOK when I got > > it. It's a nice supply with variable output voltage, current limiting, > > safety interlocks, etc. I hooked it up to the tube and let it run for a > > few weeks. Tube pumped down successfully and is working AOK now. My > guess > > is that my A18 wasn't capable of supplying the current needed to pump > down > > the tube but the external power supply was. > > > > > > Note: I proceeded very carefully with the external supply, letting it > > run for days at lower voltages and gradually working up to 3500. Also, I > > kept the current limit set as low as it would go. Anyway, it worked for > > me. Good luck with yours. > > > > > > Rodger > > > > > > -----Original Message----- > > > From: Jim Muehlberg via time-nuts <time-nuts@lists.febo.com> > > > Sent: Wednesday, November 9, 2022 11:07 AM > > > To: time-nuts@lists.febo.com > > > Cc: Jim Muehlberg <jmuehlbe@nrao.edu> > > > Subject: [time-nuts] 5061A HV Supply, et al. > > > > > > I've recently come into possession of a 5061A. It is claimed that the > > tube is exhausted, but I'm unwilling to scrap it (horrors!) yet. I'm not > > well equipped for HV measurements, but it appears the +3500 V supply is > not > > working. So, I'm hoping the tube just needs to be pumped since it hasn't > > been powered for years. Also missing the A3 assembly, but I'll worry > about > > that later... > > > > > > Anyone willing to part with their A18 or loan it to me? > > > > > > Thanks! > > > > > -- > > > > Jim Muehlberg > > Senior Engineer > > > > National Radio Astronomy Observatory > > ngVLA Local Oscillator Lead > > > > 1180 Boxwood Estates Rd B-111 > > Charlottesville, VA 22903-4602 > > P 434.296.0270 > > F 434.296.0324 > > www.cv.nrao.edu/~jmuehlbe > > _______________________________________________ > > time-nuts mailing list -- time-nuts@lists.febo.com > > To unsubscribe send an email to time-nuts-leave@lists.febo.com > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com > >
LV
Lester Veenstra
Mon, Nov 14, 2022 8:54 PM

Of course  a small “Variac” on the primary will get the voltage to what you want

Lester B Veenstra  K1YCM  MØYCM  W8YCM  6Y6Y W8YCM/6Y 6Y8LV (Reformed USNSG CTM1)

mailto:lester@veenstras.com lester@veenstras.com

452 Stable Ln

Keyser WV 26726 USA

GPS: 39.336826 N  78.982287 W (Google)

GPS: 39.33682 N  78.9823741 W (GPSDO)

Telephones:

Home:            +1-304-289-6057

US cell          +1-304-790-9192

Jamaica cell:    +1-876-456-8898

From: Andy Talbot [mailto:andy.g4jnt@gmail.com]
Sent: Monday, November 14, 2022 3:44 PM
To: Lester Veenstra
Cc: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Re: 5061A HV Supply, et al. {External}

I've got three transformers going spare ... :-)

Andy  G4JNT

IO90IV58AK12 (and others)

On Mon, 14 Nov 2022 at 20:40, Lester Veenstra m0ycm@veenstras.com wrote:

Sacrifice an old microwave oven to the Time Gods for its transformer

GPS: 39.336826 N  78.982287 W (Google)  =                    FM09MI20DU01
GPS: 39.33682 N  78.9823741 W (GPSDO) =                    FM09MI20CU70

Telephones:
Home:            +1-304-289-6057
US cell          +1-304-790-9192
Jamaica cell:    +1-876-456-8898

-----Original Message-----
From: Andy Talbot via time-nuts [mailto:time-nuts@lists.febo.com]
Sent: Monday, November 14, 2022 3:06 PM
To: Discussion of precise time and frequency measurement
Cc: Jim Muehlberg; Andy Talbot
Subject: [time-nuts] Re: 5061A HV Supply, et al. {External}

A simple flyback SMPSU with a Cockcroft-Walton tied on the end will give
you a simple high voltage supply.  A chain of high value resistors and you
can even stabilise it.

Andy
www.g4jnt.com

On Mon, 14 Nov 2022 at 20:00, Jim Muehlberg via time-nuts <
time-nuts@lists.febo.com> wrote:

Rodger,

Thanks for the encouragement!  I'm trying not to spend too much effort
on this - sort of a distraction from real work.  One would think that
NRAO would have a HV power supply lying around somewhere, but astronomy
has gone solid state now for decades!  We have a grad student trying to
nurse along another 5061 and we'd like to have a pair. I guess I'll
slice open the A18 module and see what makes it tick.

Thanks again,

Jim

On 11/10/2022 9:11 AM, Rodger via time-nuts wrote:

Hey Jim,

I got a 5061a a couple years ago with high ION pump current.  In my case

the 3500 volts was there but the tube was not pumping down, even after
being on for weeks.  I found a Terranova 741 Ion pump power supply on eBay
for under $100.  It was listed as not working but worked AOK when I got
it.  It's a nice supply with variable output voltage, current limiting,
safety interlocks, etc.  I hooked it up to the tube and let it run for a
few weeks.  Tube pumped down successfully and is working AOK now.  My

guess

is that my A18 wasn't capable of supplying the current needed to pump down
the tube but the external power supply was.

Note: I proceeded very carefully with the external supply, letting it

run for days at lower voltages and gradually working up to 3500.  Also, I
kept the current limit set as low as it would go.  Anyway, it worked for
me.  Good luck with yours.

Rodger

-----Original Message-----
From: Jim Muehlberg via time-nuts time-nuts@lists.febo.com
Sent: Wednesday, November 9, 2022 11:07 AM
To: time-nuts@lists.febo.com
Cc: Jim Muehlberg jmuehlbe@nrao.edu
Subject: [time-nuts] 5061A HV Supply, et al.

I've recently come into possession of a 5061A.  It is claimed that the

tube is exhausted, but I'm unwilling to scrap it (horrors!) yet.  I'm not
well equipped for HV measurements, but it appears the +3500 V supply is

not

working.  So, I'm hoping the tube just needs to be pumped since it hasn't
been powered for years.  Also missing the A3 assembly, but I'll worry

about

that later...

Anyone willing to part with their A18 or loan it to me?

Thanks!

--

Jim Muehlberg
Senior Engineer

National Radio Astronomy Observatory
ngVLA Local Oscillator Lead

1180 Boxwood Estates Rd B-111
Charlottesville, VA 22903-4602
P 434.296.0270
F 434.296.0324
www.cv.nrao.edu/~jmuehlbe


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

Of course a small “Variac” on the primary will get the voltage to what you want Lester B Veenstra K1YCM MØYCM W8YCM 6Y6Y W8YCM/6Y 6Y8LV (Reformed USNSG CTM1) <mailto:lester@veenstras.com> lester@veenstras.com 452 Stable Ln Keyser WV 26726 USA GPS: 39.336826 N 78.982287 W (Google) GPS: 39.33682 N 78.9823741 W (GPSDO) Telephones: Home: +1-304-289-6057 US cell +1-304-790-9192 Jamaica cell: +1-876-456-8898 From: Andy Talbot [mailto:andy.g4jnt@gmail.com] Sent: Monday, November 14, 2022 3:44 PM To: Lester Veenstra Cc: Discussion of precise time and frequency measurement Subject: Re: [time-nuts] Re: 5061A HV Supply, et al. {External} I've got three transformers going spare ... :-) Andy G4JNT IO90IV58AK12 (and others) On Mon, 14 Nov 2022 at 20:40, Lester Veenstra <m0ycm@veenstras.com> wrote: Sacrifice an old microwave oven to the Time Gods for its transformer GPS: 39.336826 N 78.982287 W (Google) = FM09MI20DU01 GPS: 39.33682 N 78.9823741 W (GPSDO) = FM09MI20CU70 Telephones: Home: +1-304-289-6057 US cell +1-304-790-9192 Jamaica cell: +1-876-456-8898 -----Original Message----- From: Andy Talbot via time-nuts [mailto:time-nuts@lists.febo.com] Sent: Monday, November 14, 2022 3:06 PM To: Discussion of precise time and frequency measurement Cc: Jim Muehlberg; Andy Talbot Subject: [time-nuts] Re: 5061A HV Supply, et al. {External} A simple flyback SMPSU with a Cockcroft-Walton tied on the end will give you a simple high voltage supply. A chain of high value resistors and you can even stabilise it. Andy www.g4jnt.com On Mon, 14 Nov 2022 at 20:00, Jim Muehlberg via time-nuts < time-nuts@lists.febo.com> wrote: > Rodger, > > Thanks for the encouragement! I'm trying not to spend too much effort > on this - sort of a distraction from real work. One would think that > NRAO would have a HV power supply lying around somewhere, but astronomy > has gone solid state now for decades! We have a grad student trying to > nurse along another 5061 and we'd like to have a pair. I guess I'll > slice open the A18 module and see what makes it tick. > > Thanks again, > > Jim > > On 11/10/2022 9:11 AM, Rodger via time-nuts wrote: > > Hey Jim, > > > > I got a 5061a a couple years ago with high ION pump current. In my case > the 3500 volts was there but the tube was not pumping down, even after > being on for weeks. I found a Terranova 741 Ion pump power supply on eBay > for under $100. It was listed as not working but worked AOK when I got > it. It's a nice supply with variable output voltage, current limiting, > safety interlocks, etc. I hooked it up to the tube and let it run for a > few weeks. Tube pumped down successfully and is working AOK now. My guess > is that my A18 wasn't capable of supplying the current needed to pump down > the tube but the external power supply was. > > > > Note: I proceeded very carefully with the external supply, letting it > run for days at lower voltages and gradually working up to 3500. Also, I > kept the current limit set as low as it would go. Anyway, it worked for > me. Good luck with yours. > > > > Rodger > > > > -----Original Message----- > > From: Jim Muehlberg via time-nuts <time-nuts@lists.febo.com> > > Sent: Wednesday, November 9, 2022 11:07 AM > > To: time-nuts@lists.febo.com > > Cc: Jim Muehlberg <jmuehlbe@nrao.edu> > > Subject: [time-nuts] 5061A HV Supply, et al. > > > > I've recently come into possession of a 5061A. It is claimed that the > tube is exhausted, but I'm unwilling to scrap it (horrors!) yet. I'm not > well equipped for HV measurements, but it appears the +3500 V supply is not > working. So, I'm hoping the tube just needs to be pumped since it hasn't > been powered for years. Also missing the A3 assembly, but I'll worry about > that later... > > > > Anyone willing to part with their A18 or loan it to me? > > > > Thanks! > > > -- > > Jim Muehlberg > Senior Engineer > > National Radio Astronomy Observatory > ngVLA Local Oscillator Lead > > 1180 Boxwood Estates Rd B-111 > Charlottesville, VA 22903-4602 > P 434.296.0270 > F 434.296.0324 > www.cv.nrao.edu/~jmuehlbe > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com _______________________________________________ time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe send an email to time-nuts-leave@lists.febo.com
JM
Jim Muehlberg
Tue, Nov 15, 2022 12:14 AM

Paul,

My understanding is that the 3500 V is on as soon as power is applied. 
Disconnected, I get 750V at the HV lead.  I first connected a 1M Ohm
resistor divider because I don't really have HV measuring stuff (10K to

  1. and it read 350 V.  Indeed the meter is pegged, but I suspect that is
    due to some internal fault in the A18 as well.  So unless there is some
    turn on signal to the 3500 V supply, I suspect it is bad.  This whole
    thing may be a basket case.  No one knows the history.

I'm also missing the divider assembly, so that may put the brakes on the
whole project, but I'll worry about that once I have attempted to pump down.

Thanks for the advice!

Jim

On 11/14/2022 3:25 PM, paul swed via time-nuts wrote:

Jim
I may be missing something here. Your suggesting the 3500 V ion pump supply
is bad.
But I believe the ion pump meter is pegged from an earlier post. If true
then the 3500 V is fine. If the tubes dirty and drawing too much current
there is a lockout circuit on a15 that will not allow the CS oven and 2500
V to come up until the ion pump is below about 15-20ua on the meter. It
senses the current being drawn by the 3500 HV. Enters a15 on pin 1 and
inhibits the rest of the system.
If my thoughts are correct then you need to pump down the tube as Rodger
says with a 3500 V supply and as I suggested 300 ua is way more than
enough. There is a chance the tube may work in a relative quality sense.
That is it could lock but there are so few Cs that the output of the system
is noisy.

Regards
Paul
WB8TSL

On Mon, Nov 14, 2022 at 3:00 PM Jim Muehlberg via time-nuts <
time-nuts@lists.febo.com> wrote:

Rodger,

Thanks for the encouragement!  I'm trying not to spend too much effort
on this - sort of a distraction from real work.  One would think that
NRAO would have a HV power supply lying around somewhere, but astronomy
has gone solid state now for decades!  We have a grad student trying to
nurse along another 5061 and we'd like to have a pair. I guess I'll
slice open the A18 module and see what makes it tick.

Thanks again,

Jim

On 11/10/2022 9:11 AM, Rodger via time-nuts wrote:

Hey Jim,

I got a 5061a a couple years ago with high ION pump current.  In my case

the 3500 volts was there but the tube was not pumping down, even after
being on for weeks.  I found a Terranova 741 Ion pump power supply on eBay
for under $100.  It was listed as not working but worked AOK when I got
it.  It's a nice supply with variable output voltage, current limiting,
safety interlocks, etc.  I hooked it up to the tube and let it run for a
few weeks.  Tube pumped down successfully and is working AOK now.  My guess
is that my A18 wasn't capable of supplying the current needed to pump down
the tube but the external power supply was.

Note: I proceeded very carefully with the external supply, letting it

run for days at lower voltages and gradually working up to 3500.  Also, I
kept the current limit set as low as it would go.  Anyway, it worked for
me.  Good luck with yours.

Rodger

-----Original Message-----
From: Jim Muehlberg via time-nuts time-nuts@lists.febo.com
Sent: Wednesday, November 9, 2022 11:07 AM
To: time-nuts@lists.febo.com
Cc: Jim Muehlberg jmuehlbe@nrao.edu
Subject: [time-nuts] 5061A HV Supply, et al.

I've recently come into possession of a 5061A.  It is claimed that the

tube is exhausted, but I'm unwilling to scrap it (horrors!) yet.  I'm not
well equipped for HV measurements, but it appears the +3500 V supply is not
working.  So, I'm hoping the tube just needs to be pumped since it hasn't
been powered for years.  Also missing the A3 assembly, but I'll worry about
that later...

Anyone willing to part with their A18 or loan it to me?

Thanks!

--

Jim Muehlberg
Senior Engineer

National Radio Astronomy Observatory
ngVLA Local Oscillator Lead

1180 Boxwood Estates Rd B-111
Charlottesville, VA 22903-4602
P 434.296.0270
F 434.296.0324
www.cv.nrao.edu/~jmuehlbe


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

--

Jim Muehlberg
Senior Engineer

National Radio Astronomy Observatory
ngVLA Local Oscillator Lead

1180 Boxwood Estates Rd B-111
Charlottesville, VA 22903-4602
P 434.296.0270
F 434.296.0324
www.cv.nrao.edu/~jmuehlbe

Paul, My understanding is that the 3500 V is on as soon as power is applied.  Disconnected, I get 750V at the HV lead.  I first connected a 1M Ohm resistor divider because I don't really have HV measuring stuff (10K to 1) and it read 350 V.  Indeed the meter is pegged, but I suspect that is due to some internal fault in the A18 as well.  So unless there is some turn on signal to the 3500 V supply, I suspect it is bad.  This whole thing may be a basket case.  No one knows the history. I'm also missing the divider assembly, so that may put the brakes on the whole project, but I'll worry about that once I have attempted to pump down. Thanks for the advice! Jim On 11/14/2022 3:25 PM, paul swed via time-nuts wrote: > Jim > I may be missing something here. Your suggesting the 3500 V ion pump supply > is bad. > But I believe the ion pump meter is pegged from an earlier post. If true > then the 3500 V is fine. If the tubes dirty and drawing too much current > there is a lockout circuit on a15 that will not allow the CS oven and 2500 > V to come up until the ion pump is below about 15-20ua on the meter. It > senses the current being drawn by the 3500 HV. Enters a15 on pin 1 and > inhibits the rest of the system. > If my thoughts are correct then you need to pump down the tube as Rodger > says with a 3500 V supply and as I suggested 300 ua is way more than > enough. There is a chance the tube may work in a relative quality sense. > That is it could lock but there are so few Cs that the output of the system > is noisy. > > Regards > Paul > WB8TSL > > > On Mon, Nov 14, 2022 at 3:00 PM Jim Muehlberg via time-nuts < > time-nuts@lists.febo.com> wrote: > >> Rodger, >> >> Thanks for the encouragement! I'm trying not to spend too much effort >> on this - sort of a distraction from real work. One would think that >> NRAO would have a HV power supply lying around somewhere, but astronomy >> has gone solid state now for decades! We have a grad student trying to >> nurse along another 5061 and we'd like to have a pair. I guess I'll >> slice open the A18 module and see what makes it tick. >> >> Thanks again, >> >> Jim >> >> On 11/10/2022 9:11 AM, Rodger via time-nuts wrote: >>> Hey Jim, >>> >>> I got a 5061a a couple years ago with high ION pump current. In my case >> the 3500 volts was there but the tube was not pumping down, even after >> being on for weeks. I found a Terranova 741 Ion pump power supply on eBay >> for under $100. It was listed as not working but worked AOK when I got >> it. It's a nice supply with variable output voltage, current limiting, >> safety interlocks, etc. I hooked it up to the tube and let it run for a >> few weeks. Tube pumped down successfully and is working AOK now. My guess >> is that my A18 wasn't capable of supplying the current needed to pump down >> the tube but the external power supply was. >>> Note: I proceeded very carefully with the external supply, letting it >> run for days at lower voltages and gradually working up to 3500. Also, I >> kept the current limit set as low as it would go. Anyway, it worked for >> me. Good luck with yours. >>> Rodger >>> >>> -----Original Message----- >>> From: Jim Muehlberg via time-nuts <time-nuts@lists.febo.com> >>> Sent: Wednesday, November 9, 2022 11:07 AM >>> To: time-nuts@lists.febo.com >>> Cc: Jim Muehlberg <jmuehlbe@nrao.edu> >>> Subject: [time-nuts] 5061A HV Supply, et al. >>> >>> I've recently come into possession of a 5061A. It is claimed that the >> tube is exhausted, but I'm unwilling to scrap it (horrors!) yet. I'm not >> well equipped for HV measurements, but it appears the +3500 V supply is not >> working. So, I'm hoping the tube just needs to be pumped since it hasn't >> been powered for years. Also missing the A3 assembly, but I'll worry about >> that later... >>> Anyone willing to part with their A18 or loan it to me? >>> >>> Thanks! >>> >> -- >> >> Jim Muehlberg >> Senior Engineer >> >> National Radio Astronomy Observatory >> ngVLA Local Oscillator Lead >> >> 1180 Boxwood Estates Rd B-111 >> Charlottesville, VA 22903-4602 >> P 434.296.0270 >> F 434.296.0324 >> www.cv.nrao.edu/~jmuehlbe >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe send an email to time-nuts-leave@lists.febo.com > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com -- Jim Muehlberg Senior Engineer National Radio Astronomy Observatory ngVLA Local Oscillator Lead 1180 Boxwood Estates Rd B-111 Charlottesville, VA 22903-4602 P 434.296.0270 F 434.296.0324 www.cv.nrao.edu/~jmuehlbe
PS
paul swed
Tue, Nov 15, 2022 2:27 AM

Jim
OK I thought the thread said the meter is pegged and yes at turn on the ion
pump 3500 V is always on. The switching and protection I am speaking about
is for the other circuits. They will not start if the pump current is to
high. The ION pump sends a signal that inhibits their start. A 1M R is far
too low of a resistance its loading the supply down. Try something like
40-60 Meg. Maybe a string of 10 M with a 100k R at the bottom. Measure the
V across the 100K. In normal operation the 3500 V will be drawing 4-14ua if
the tubes reasonably pumped down.
So thinking about this. There is a twist HV connector on the chassis that
the 3500 V goes through. I am reluctant to say this but if that were off
and far away from anything to arc to and the ion pump meter went to close
to zero then you would know the supply is most likely working and you have
a very dirty tube. I just do not like the thought of a 3500 V plug floating
free.
Regards
Paul
WB8TSL

On Mon, Nov 14, 2022 at 7:51 PM Jim Muehlberg via time-nuts <
time-nuts@lists.febo.com> wrote:

Paul,

My understanding is that the 3500 V is on as soon as power is applied.
Disconnected, I get 750V at the HV lead.  I first connected a 1M Ohm
resistor divider because I don't really have HV measuring stuff (10K to

  1. and it read 350 V.  Indeed the meter is pegged, but I suspect that is
    due to some internal fault in the A18 as well.  So unless there is some
    turn on signal to the 3500 V supply, I suspect it is bad.  This whole
    thing may be a basket case.  No one knows the history.

I'm also missing the divider assembly, so that may put the brakes on the
whole project, but I'll worry about that once I have attempted to pump
down.

Thanks for the advice!

Jim

On 11/14/2022 3:25 PM, paul swed via time-nuts wrote:

Jim
I may be missing something here. Your suggesting the 3500 V ion pump

supply

is bad.
But I believe the ion pump meter is pegged from an earlier post. If true
then the 3500 V is fine. If the tubes dirty and drawing too much current
there is a lockout circuit on a15 that will not allow the CS oven and

2500

V to come up until the ion pump is below about 15-20ua on the meter. It
senses the current being drawn by the 3500 HV. Enters a15 on pin 1 and
inhibits the rest of the system.
If my thoughts are correct then you need to pump down the tube as Rodger
says with a 3500 V supply and as I suggested 300 ua is way more than
enough. There is a chance the tube may work in a relative quality sense.
That is it could lock but there are so few Cs that the output of the

system

is noisy.

Regards
Paul
WB8TSL

On Mon, Nov 14, 2022 at 3:00 PM Jim Muehlberg via time-nuts <
time-nuts@lists.febo.com> wrote:

Rodger,

Thanks for the encouragement!  I'm trying not to spend too much effort
on this - sort of a distraction from real work.  One would think that
NRAO would have a HV power supply lying around somewhere, but astronomy
has gone solid state now for decades!  We have a grad student trying to
nurse along another 5061 and we'd like to have a pair. I guess I'll
slice open the A18 module and see what makes it tick.

Thanks again,

Jim

On 11/10/2022 9:11 AM, Rodger via time-nuts wrote:

Hey Jim,

I got a 5061a a couple years ago with high ION pump current.  In my

case

the 3500 volts was there but the tube was not pumping down, even after
being on for weeks.  I found a Terranova 741 Ion pump power supply on

eBay

for under $100.  It was listed as not working but worked AOK when I got
it.  It's a nice supply with variable output voltage, current limiting,
safety interlocks, etc.  I hooked it up to the tube and let it run for a
few weeks.  Tube pumped down successfully and is working AOK now.  My

guess

is that my A18 wasn't capable of supplying the current needed to pump

down

the tube but the external power supply was.

Note: I proceeded very carefully with the external supply, letting it

run for days at lower voltages and gradually working up to 3500.  Also,

I

kept the current limit set as low as it would go.  Anyway, it worked for
me.  Good luck with yours.

Rodger

-----Original Message-----
From: Jim Muehlberg via time-nuts time-nuts@lists.febo.com
Sent: Wednesday, November 9, 2022 11:07 AM
To: time-nuts@lists.febo.com
Cc: Jim Muehlberg jmuehlbe@nrao.edu
Subject: [time-nuts] 5061A HV Supply, et al.

I've recently come into possession of a 5061A.  It is claimed that the

tube is exhausted, but I'm unwilling to scrap it (horrors!) yet.  I'm

not

well equipped for HV measurements, but it appears the +3500 V supply is

not

working.  So, I'm hoping the tube just needs to be pumped since it

hasn't

been powered for years.  Also missing the A3 assembly, but I'll worry

about

that later...

Anyone willing to part with their A18 or loan it to me?

Thanks!

--

Jim Muehlberg
Senior Engineer

National Radio Astronomy Observatory
ngVLA Local Oscillator Lead

1180 Boxwood Estates Rd B-111
Charlottesville, VA 22903-4602
P 434.296.0270
F 434.296.0324
www.cv.nrao.edu/~jmuehlbe


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

--

Jim Muehlberg
Senior Engineer

National Radio Astronomy Observatory
ngVLA Local Oscillator Lead

1180 Boxwood Estates Rd B-111
Charlottesville, VA 22903-4602
P 434.296.0270
F 434.296.0324
www.cv.nrao.edu/~jmuehlbe


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

Jim OK I thought the thread said the meter is pegged and yes at turn on the ion pump 3500 V is always on. The switching and protection I am speaking about is for the other circuits. They will not start if the pump current is to high. The ION pump sends a signal that inhibits their start. A 1M R is far too low of a resistance its loading the supply down. Try something like 40-60 Meg. Maybe a string of 10 M with a 100k R at the bottom. Measure the V across the 100K. In normal operation the 3500 V will be drawing 4-14ua if the tubes reasonably pumped down. So thinking about this. There is a twist HV connector on the chassis that the 3500 V goes through. I am reluctant to say this but if that were off and far away from anything to arc to and the ion pump meter went to close to zero then you would know the supply is most likely working and you have a very dirty tube. I just do not like the thought of a 3500 V plug floating free. Regards Paul WB8TSL On Mon, Nov 14, 2022 at 7:51 PM Jim Muehlberg via time-nuts < time-nuts@lists.febo.com> wrote: > Paul, > > My understanding is that the 3500 V is on as soon as power is applied. > Disconnected, I get 750V at the HV lead. I first connected a 1M Ohm > resistor divider because I don't really have HV measuring stuff (10K to > 1) and it read 350 V. Indeed the meter is pegged, but I suspect that is > due to some internal fault in the A18 as well. So unless there is some > turn on signal to the 3500 V supply, I suspect it is bad. This whole > thing may be a basket case. No one knows the history. > > I'm also missing the divider assembly, so that may put the brakes on the > whole project, but I'll worry about that once I have attempted to pump > down. > > Thanks for the advice! > > Jim > > On 11/14/2022 3:25 PM, paul swed via time-nuts wrote: > > Jim > > I may be missing something here. Your suggesting the 3500 V ion pump > supply > > is bad. > > But I believe the ion pump meter is pegged from an earlier post. If true > > then the 3500 V is fine. If the tubes dirty and drawing too much current > > there is a lockout circuit on a15 that will not allow the CS oven and > 2500 > > V to come up until the ion pump is below about 15-20ua on the meter. It > > senses the current being drawn by the 3500 HV. Enters a15 on pin 1 and > > inhibits the rest of the system. > > If my thoughts are correct then you need to pump down the tube as Rodger > > says with a 3500 V supply and as I suggested 300 ua is way more than > > enough. There is a chance the tube may work in a relative quality sense. > > That is it could lock but there are so few Cs that the output of the > system > > is noisy. > > > > Regards > > Paul > > WB8TSL > > > > > > On Mon, Nov 14, 2022 at 3:00 PM Jim Muehlberg via time-nuts < > > time-nuts@lists.febo.com> wrote: > > > >> Rodger, > >> > >> Thanks for the encouragement! I'm trying not to spend too much effort > >> on this - sort of a distraction from real work. One would think that > >> NRAO would have a HV power supply lying around somewhere, but astronomy > >> has gone solid state now for decades! We have a grad student trying to > >> nurse along another 5061 and we'd like to have a pair. I guess I'll > >> slice open the A18 module and see what makes it tick. > >> > >> Thanks again, > >> > >> Jim > >> > >> On 11/10/2022 9:11 AM, Rodger via time-nuts wrote: > >>> Hey Jim, > >>> > >>> I got a 5061a a couple years ago with high ION pump current. In my > case > >> the 3500 volts was there but the tube was not pumping down, even after > >> being on for weeks. I found a Terranova 741 Ion pump power supply on > eBay > >> for under $100. It was listed as not working but worked AOK when I got > >> it. It's a nice supply with variable output voltage, current limiting, > >> safety interlocks, etc. I hooked it up to the tube and let it run for a > >> few weeks. Tube pumped down successfully and is working AOK now. My > guess > >> is that my A18 wasn't capable of supplying the current needed to pump > down > >> the tube but the external power supply was. > >>> Note: I proceeded very carefully with the external supply, letting it > >> run for days at lower voltages and gradually working up to 3500. Also, > I > >> kept the current limit set as low as it would go. Anyway, it worked for > >> me. Good luck with yours. > >>> Rodger > >>> > >>> -----Original Message----- > >>> From: Jim Muehlberg via time-nuts <time-nuts@lists.febo.com> > >>> Sent: Wednesday, November 9, 2022 11:07 AM > >>> To: time-nuts@lists.febo.com > >>> Cc: Jim Muehlberg <jmuehlbe@nrao.edu> > >>> Subject: [time-nuts] 5061A HV Supply, et al. > >>> > >>> I've recently come into possession of a 5061A. It is claimed that the > >> tube is exhausted, but I'm unwilling to scrap it (horrors!) yet. I'm > not > >> well equipped for HV measurements, but it appears the +3500 V supply is > not > >> working. So, I'm hoping the tube just needs to be pumped since it > hasn't > >> been powered for years. Also missing the A3 assembly, but I'll worry > about > >> that later... > >>> Anyone willing to part with their A18 or loan it to me? > >>> > >>> Thanks! > >>> > >> -- > >> > >> Jim Muehlberg > >> Senior Engineer > >> > >> National Radio Astronomy Observatory > >> ngVLA Local Oscillator Lead > >> > >> 1180 Boxwood Estates Rd B-111 > >> Charlottesville, VA 22903-4602 > >> P 434.296.0270 > >> F 434.296.0324 > >> www.cv.nrao.edu/~jmuehlbe > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@lists.febo.com > >> To unsubscribe send an email to time-nuts-leave@lists.febo.com > > _______________________________________________ > > time-nuts mailing list -- time-nuts@lists.febo.com > > To unsubscribe send an email to time-nuts-leave@lists.febo.com > -- > > Jim Muehlberg > Senior Engineer > > National Radio Astronomy Observatory > ngVLA Local Oscillator Lead > > 1180 Boxwood Estates Rd B-111 > Charlottesville, VA 22903-4602 > P 434.296.0270 > F 434.296.0324 > www.cv.nrao.edu/~jmuehlbe > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com
CH
Chris Hastreiter
Tue, Nov 15, 2022 2:30 AM

Jim,

The 3500V supply is designed to be a soft supply that folds under load. A
1MOhm load is enough to make it collapse down to 750V. The meter display
estimates current by measuring how much the voltage sags.

If you want to measure open circuit voltage you'll need to get closer to a
1GOhm measurement impedance. If you want to crosscheck the current through
the ion pump you can unmount the tube and connect an ammeter between it and
the chassis (the return path for the ion pump) and measure the current
directly.

Thanks,

Chris

On Mon, Nov 14, 2022, 6:51 PM Jim Muehlberg via time-nuts <
time-nuts@lists.febo.com> wrote:

Paul,

My understanding is that the 3500 V is on as soon as power is applied.
Disconnected, I get 750V at the HV lead.  I first connected a 1M Ohm
resistor divider because I don't really have HV measuring stuff (10K to

  1. and it read 350 V.  Indeed the meter is pegged, but I suspect that is
    due to some internal fault in the A18 as well.  So unless there is some
    turn on signal to the 3500 V supply, I suspect it is bad.  This whole
    thing may be a basket case.  No one knows the history.

I'm also missing the divider assembly, so that may put the brakes on the
whole project, but I'll worry about that once I have attempted to pump
down.

Thanks for the advice!

Jim

On 11/14/2022 3:25 PM, paul swed via time-nuts wrote:

Jim
I may be missing something here. Your suggesting the 3500 V ion pump

supply

is bad.
But I believe the ion pump meter is pegged from an earlier post. If true
then the 3500 V is fine. If the tubes dirty and drawing too much current
there is a lockout circuit on a15 that will not allow the CS oven and

2500

V to come up until the ion pump is below about 15-20ua on the meter. It
senses the current being drawn by the 3500 HV. Enters a15 on pin 1 and
inhibits the rest of the system.
If my thoughts are correct then you need to pump down the tube as Rodger
says with a 3500 V supply and as I suggested 300 ua is way more than
enough. There is a chance the tube may work in a relative quality sense.
That is it could lock but there are so few Cs that the output of the

system

is noisy.

Regards
Paul
WB8TSL

On Mon, Nov 14, 2022 at 3:00 PM Jim Muehlberg via time-nuts <
time-nuts@lists.febo.com> wrote:

Rodger,

Thanks for the encouragement!  I'm trying not to spend too much effort
on this - sort of a distraction from real work.  One would think that
NRAO would have a HV power supply lying around somewhere, but astronomy
has gone solid state now for decades!  We have a grad student trying to
nurse along another 5061 and we'd like to have a pair. I guess I'll
slice open the A18 module and see what makes it tick.

Thanks again,

Jim

On 11/10/2022 9:11 AM, Rodger via time-nuts wrote:

Hey Jim,

I got a 5061a a couple years ago with high ION pump current.  In my

case

the 3500 volts was there but the tube was not pumping down, even after
being on for weeks.  I found a Terranova 741 Ion pump power supply on

eBay

for under $100.  It was listed as not working but worked AOK when I got
it.  It's a nice supply with variable output voltage, current limiting,
safety interlocks, etc.  I hooked it up to the tube and let it run for a
few weeks.  Tube pumped down successfully and is working AOK now.  My

guess

is that my A18 wasn't capable of supplying the current needed to pump

down

the tube but the external power supply was.

Note: I proceeded very carefully with the external supply, letting it

run for days at lower voltages and gradually working up to 3500.  Also,

I

kept the current limit set as low as it would go.  Anyway, it worked for
me.  Good luck with yours.

Rodger

-----Original Message-----
From: Jim Muehlberg via time-nuts time-nuts@lists.febo.com
Sent: Wednesday, November 9, 2022 11:07 AM
To: time-nuts@lists.febo.com
Cc: Jim Muehlberg jmuehlbe@nrao.edu
Subject: [time-nuts] 5061A HV Supply, et al.

I've recently come into possession of a 5061A.  It is claimed that the

tube is exhausted, but I'm unwilling to scrap it (horrors!) yet.  I'm

not

well equipped for HV measurements, but it appears the +3500 V supply is

not

working.  So, I'm hoping the tube just needs to be pumped since it

hasn't

been powered for years.  Also missing the A3 assembly, but I'll worry

about

that later...

Anyone willing to part with their A18 or loan it to me?

Thanks!

--

Jim Muehlberg
Senior Engineer

National Radio Astronomy Observatory
ngVLA Local Oscillator Lead

1180 Boxwood Estates Rd B-111
Charlottesville, VA 22903-4602
P 434.296.0270
F 434.296.0324
www.cv.nrao.edu/~jmuehlbe


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

--

Jim Muehlberg
Senior Engineer

National Radio Astronomy Observatory
ngVLA Local Oscillator Lead

1180 Boxwood Estates Rd B-111
Charlottesville, VA 22903-4602
P 434.296.0270
F 434.296.0324
www.cv.nrao.edu/~jmuehlbe


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

Jim, The 3500V supply is designed to be a soft supply that folds under load. A 1MOhm load is enough to make it collapse down to 750V. The meter display estimates current by measuring how much the voltage sags. If you want to measure open circuit voltage you'll need to get closer to a 1GOhm measurement impedance. If you want to crosscheck the current through the ion pump you can unmount the tube and connect an ammeter between it and the chassis (the return path for the ion pump) and measure the current directly. Thanks, Chris On Mon, Nov 14, 2022, 6:51 PM Jim Muehlberg via time-nuts < time-nuts@lists.febo.com> wrote: > Paul, > > My understanding is that the 3500 V is on as soon as power is applied. > Disconnected, I get 750V at the HV lead. I first connected a 1M Ohm > resistor divider because I don't really have HV measuring stuff (10K to > 1) and it read 350 V. Indeed the meter is pegged, but I suspect that is > due to some internal fault in the A18 as well. So unless there is some > turn on signal to the 3500 V supply, I suspect it is bad. This whole > thing may be a basket case. No one knows the history. > > I'm also missing the divider assembly, so that may put the brakes on the > whole project, but I'll worry about that once I have attempted to pump > down. > > Thanks for the advice! > > Jim > > On 11/14/2022 3:25 PM, paul swed via time-nuts wrote: > > Jim > > I may be missing something here. Your suggesting the 3500 V ion pump > supply > > is bad. > > But I believe the ion pump meter is pegged from an earlier post. If true > > then the 3500 V is fine. If the tubes dirty and drawing too much current > > there is a lockout circuit on a15 that will not allow the CS oven and > 2500 > > V to come up until the ion pump is below about 15-20ua on the meter. It > > senses the current being drawn by the 3500 HV. Enters a15 on pin 1 and > > inhibits the rest of the system. > > If my thoughts are correct then you need to pump down the tube as Rodger > > says with a 3500 V supply and as I suggested 300 ua is way more than > > enough. There is a chance the tube may work in a relative quality sense. > > That is it could lock but there are so few Cs that the output of the > system > > is noisy. > > > > Regards > > Paul > > WB8TSL > > > > > > On Mon, Nov 14, 2022 at 3:00 PM Jim Muehlberg via time-nuts < > > time-nuts@lists.febo.com> wrote: > > > >> Rodger, > >> > >> Thanks for the encouragement! I'm trying not to spend too much effort > >> on this - sort of a distraction from real work. One would think that > >> NRAO would have a HV power supply lying around somewhere, but astronomy > >> has gone solid state now for decades! We have a grad student trying to > >> nurse along another 5061 and we'd like to have a pair. I guess I'll > >> slice open the A18 module and see what makes it tick. > >> > >> Thanks again, > >> > >> Jim > >> > >> On 11/10/2022 9:11 AM, Rodger via time-nuts wrote: > >>> Hey Jim, > >>> > >>> I got a 5061a a couple years ago with high ION pump current. In my > case > >> the 3500 volts was there but the tube was not pumping down, even after > >> being on for weeks. I found a Terranova 741 Ion pump power supply on > eBay > >> for under $100. It was listed as not working but worked AOK when I got > >> it. It's a nice supply with variable output voltage, current limiting, > >> safety interlocks, etc. I hooked it up to the tube and let it run for a > >> few weeks. Tube pumped down successfully and is working AOK now. My > guess > >> is that my A18 wasn't capable of supplying the current needed to pump > down > >> the tube but the external power supply was. > >>> Note: I proceeded very carefully with the external supply, letting it > >> run for days at lower voltages and gradually working up to 3500. Also, > I > >> kept the current limit set as low as it would go. Anyway, it worked for > >> me. Good luck with yours. > >>> Rodger > >>> > >>> -----Original Message----- > >>> From: Jim Muehlberg via time-nuts <time-nuts@lists.febo.com> > >>> Sent: Wednesday, November 9, 2022 11:07 AM > >>> To: time-nuts@lists.febo.com > >>> Cc: Jim Muehlberg <jmuehlbe@nrao.edu> > >>> Subject: [time-nuts] 5061A HV Supply, et al. > >>> > >>> I've recently come into possession of a 5061A. It is claimed that the > >> tube is exhausted, but I'm unwilling to scrap it (horrors!) yet. I'm > not > >> well equipped for HV measurements, but it appears the +3500 V supply is > not > >> working. So, I'm hoping the tube just needs to be pumped since it > hasn't > >> been powered for years. Also missing the A3 assembly, but I'll worry > about > >> that later... > >>> Anyone willing to part with their A18 or loan it to me? > >>> > >>> Thanks! > >>> > >> -- > >> > >> Jim Muehlberg > >> Senior Engineer > >> > >> National Radio Astronomy Observatory > >> ngVLA Local Oscillator Lead > >> > >> 1180 Boxwood Estates Rd B-111 > >> Charlottesville, VA 22903-4602 > >> P 434.296.0270 > >> F 434.296.0324 > >> www.cv.nrao.edu/~jmuehlbe > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@lists.febo.com > >> To unsubscribe send an email to time-nuts-leave@lists.febo.com > > _______________________________________________ > > time-nuts mailing list -- time-nuts@lists.febo.com > > To unsubscribe send an email to time-nuts-leave@lists.febo.com > -- > > Jim Muehlberg > Senior Engineer > > National Radio Astronomy Observatory > ngVLA Local Oscillator Lead > > 1180 Boxwood Estates Rd B-111 > Charlottesville, VA 22903-4602 > P 434.296.0270 > F 434.296.0324 > www.cv.nrao.edu/~jmuehlbe > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com
PS
paul swed
Tue, Nov 15, 2022 2:35 AM

Here is a pix of a unit I was testing.
[image: DSC04636.JPG]

The current meter is in the lower leg of the return. You can see the high
voltage probe on the right. I recall its 68 M and feeds a fluke DVM. The
amp meter is measuring the current that the probe is consuming. The other
connections are ground and +18.7V.
The precision voltage divider in the background was not used.
Testing the supply like this is fairly easy and just look at the schematics
to understand how to hook everything up.
Regards
Paul
WB8TSL

On Mon, Nov 14, 2022 at 9:27 PM paul swed paulswedb@gmail.com wrote:

Jim
OK I thought the thread said the meter is pegged and yes at turn on the
ion pump 3500 V is always on. The switching and protection I am speaking
about is for the other circuits. They will not start if the pump current is
to high. The ION pump sends a signal that inhibits their start. A 1M R is
far too low of a resistance its loading the supply down. Try something like
40-60 Meg. Maybe a string of 10 M with a 100k R at the bottom. Measure the
V across the 100K. In normal operation the 3500 V will be drawing 4-14ua if
the tubes reasonably pumped down.
So thinking about this. There is a twist HV connector on the chassis that
the 3500 V goes through. I am reluctant to say this but if that were off
and far away from anything to arc to and the ion pump meter went to close
to zero then you would know the supply is most likely working and you have
a very dirty tube. I just do not like the thought of a 3500 V plug floating
free.
Regards
Paul
WB8TSL

On Mon, Nov 14, 2022 at 7:51 PM Jim Muehlberg via time-nuts <
time-nuts@lists.febo.com> wrote:

Paul,

My understanding is that the 3500 V is on as soon as power is applied.
Disconnected, I get 750V at the HV lead.  I first connected a 1M Ohm
resistor divider because I don't really have HV measuring stuff (10K to

  1. and it read 350 V.  Indeed the meter is pegged, but I suspect that is
    due to some internal fault in the A18 as well.  So unless there is some
    turn on signal to the 3500 V supply, I suspect it is bad.  This whole
    thing may be a basket case.  No one knows the history.

I'm also missing the divider assembly, so that may put the brakes on the
whole project, but I'll worry about that once I have attempted to pump
down.

Thanks for the advice!

Jim

On 11/14/2022 3:25 PM, paul swed via time-nuts wrote:

Jim
I may be missing something here. Your suggesting the 3500 V ion pump

supply

is bad.
But I believe the ion pump meter is pegged from an earlier post. If true
then the 3500 V is fine. If the tubes dirty and drawing too much current
there is a lockout circuit on a15 that will not allow the CS oven and

2500

V to come up until the ion pump is below about 15-20ua on the meter. It
senses the current being drawn by the 3500 HV. Enters a15 on pin 1 and
inhibits the rest of the system.
If my thoughts are correct then you need to pump down the tube as Rodger
says with a 3500 V supply and as I suggested 300 ua is way more than
enough. There is a chance the tube may work in a relative quality sense.
That is it could lock but there are so few Cs that the output of the

system

is noisy.

Regards
Paul
WB8TSL

On Mon, Nov 14, 2022 at 3:00 PM Jim Muehlberg via time-nuts <
time-nuts@lists.febo.com> wrote:

Rodger,

Thanks for the encouragement!  I'm trying not to spend too much effort
on this - sort of a distraction from real work.  One would think that
NRAO would have a HV power supply lying around somewhere, but astronomy
has gone solid state now for decades!  We have a grad student trying to
nurse along another 5061 and we'd like to have a pair. I guess I'll
slice open the A18 module and see what makes it tick.

Thanks again,

Jim

On 11/10/2022 9:11 AM, Rodger via time-nuts wrote:

Hey Jim,

I got a 5061a a couple years ago with high ION pump current.  In my

case

the 3500 volts was there but the tube was not pumping down, even after
being on for weeks.  I found a Terranova 741 Ion pump power supply on

eBay

for under $100.  It was listed as not working but worked AOK when I got
it.  It's a nice supply with variable output voltage, current limiting,
safety interlocks, etc.  I hooked it up to the tube and let it run for

a

few weeks.  Tube pumped down successfully and is working AOK now.  My

guess

is that my A18 wasn't capable of supplying the current needed to pump

down

the tube but the external power supply was.

Note: I proceeded very carefully with the external supply, letting it

run for days at lower voltages and gradually working up to 3500.

Also, I

kept the current limit set as low as it would go.  Anyway, it worked

for

me.  Good luck with yours.

Rodger

-----Original Message-----
From: Jim Muehlberg via time-nuts time-nuts@lists.febo.com
Sent: Wednesday, November 9, 2022 11:07 AM
To: time-nuts@lists.febo.com
Cc: Jim Muehlberg jmuehlbe@nrao.edu
Subject: [time-nuts] 5061A HV Supply, et al.

I've recently come into possession of a 5061A.  It is claimed that the

tube is exhausted, but I'm unwilling to scrap it (horrors!) yet.  I'm

not

well equipped for HV measurements, but it appears the +3500 V supply

is not

working.  So, I'm hoping the tube just needs to be pumped since it

hasn't

been powered for years.  Also missing the A3 assembly, but I'll worry

about

that later...

Anyone willing to part with their A18 or loan it to me?

Thanks!

--

Jim Muehlberg
Senior Engineer

National Radio Astronomy Observatory
ngVLA Local Oscillator Lead

1180 Boxwood Estates Rd B-111
Charlottesville, VA 22903-4602
P 434.296.0270
F 434.296.0324
www.cv.nrao.edu/~jmuehlbe


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

--

Jim Muehlberg
Senior Engineer

National Radio Astronomy Observatory
ngVLA Local Oscillator Lead

1180 Boxwood Estates Rd B-111
Charlottesville, VA 22903-4602
P 434.296.0270
F 434.296.0324
www.cv.nrao.edu/~jmuehlbe


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

Here is a pix of a unit I was testing. [image: DSC04636.JPG] The current meter is in the lower leg of the return. You can see the high voltage probe on the right. I recall its 68 M and feeds a fluke DVM. The amp meter is measuring the current that the probe is consuming. The other connections are ground and +18.7V. The precision voltage divider in the background was not used. Testing the supply like this is fairly easy and just look at the schematics to understand how to hook everything up. Regards Paul WB8TSL On Mon, Nov 14, 2022 at 9:27 PM paul swed <paulswedb@gmail.com> wrote: > Jim > OK I thought the thread said the meter is pegged and yes at turn on the > ion pump 3500 V is always on. The switching and protection I am speaking > about is for the other circuits. They will not start if the pump current is > to high. The ION pump sends a signal that inhibits their start. A 1M R is > far too low of a resistance its loading the supply down. Try something like > 40-60 Meg. Maybe a string of 10 M with a 100k R at the bottom. Measure the > V across the 100K. In normal operation the 3500 V will be drawing 4-14ua if > the tubes reasonably pumped down. > So thinking about this. There is a twist HV connector on the chassis that > the 3500 V goes through. I am reluctant to say this but if that were off > and far away from anything to arc to and the ion pump meter went to close > to zero then you would know the supply is most likely working and you have > a very dirty tube. I just do not like the thought of a 3500 V plug floating > free. > Regards > Paul > WB8TSL > > > On Mon, Nov 14, 2022 at 7:51 PM Jim Muehlberg via time-nuts < > time-nuts@lists.febo.com> wrote: > >> Paul, >> >> My understanding is that the 3500 V is on as soon as power is applied. >> Disconnected, I get 750V at the HV lead. I first connected a 1M Ohm >> resistor divider because I don't really have HV measuring stuff (10K to >> 1) and it read 350 V. Indeed the meter is pegged, but I suspect that is >> due to some internal fault in the A18 as well. So unless there is some >> turn on signal to the 3500 V supply, I suspect it is bad. This whole >> thing may be a basket case. No one knows the history. >> >> I'm also missing the divider assembly, so that may put the brakes on the >> whole project, but I'll worry about that once I have attempted to pump >> down. >> >> Thanks for the advice! >> >> Jim >> >> On 11/14/2022 3:25 PM, paul swed via time-nuts wrote: >> > Jim >> > I may be missing something here. Your suggesting the 3500 V ion pump >> supply >> > is bad. >> > But I believe the ion pump meter is pegged from an earlier post. If true >> > then the 3500 V is fine. If the tubes dirty and drawing too much current >> > there is a lockout circuit on a15 that will not allow the CS oven and >> 2500 >> > V to come up until the ion pump is below about 15-20ua on the meter. It >> > senses the current being drawn by the 3500 HV. Enters a15 on pin 1 and >> > inhibits the rest of the system. >> > If my thoughts are correct then you need to pump down the tube as Rodger >> > says with a 3500 V supply and as I suggested 300 ua is way more than >> > enough. There is a chance the tube may work in a relative quality sense. >> > That is it could lock but there are so few Cs that the output of the >> system >> > is noisy. >> > >> > Regards >> > Paul >> > WB8TSL >> > >> > >> > On Mon, Nov 14, 2022 at 3:00 PM Jim Muehlberg via time-nuts < >> > time-nuts@lists.febo.com> wrote: >> > >> >> Rodger, >> >> >> >> Thanks for the encouragement! I'm trying not to spend too much effort >> >> on this - sort of a distraction from real work. One would think that >> >> NRAO would have a HV power supply lying around somewhere, but astronomy >> >> has gone solid state now for decades! We have a grad student trying to >> >> nurse along another 5061 and we'd like to have a pair. I guess I'll >> >> slice open the A18 module and see what makes it tick. >> >> >> >> Thanks again, >> >> >> >> Jim >> >> >> >> On 11/10/2022 9:11 AM, Rodger via time-nuts wrote: >> >>> Hey Jim, >> >>> >> >>> I got a 5061a a couple years ago with high ION pump current. In my >> case >> >> the 3500 volts was there but the tube was not pumping down, even after >> >> being on for weeks. I found a Terranova 741 Ion pump power supply on >> eBay >> >> for under $100. It was listed as not working but worked AOK when I got >> >> it. It's a nice supply with variable output voltage, current limiting, >> >> safety interlocks, etc. I hooked it up to the tube and let it run for >> a >> >> few weeks. Tube pumped down successfully and is working AOK now. My >> guess >> >> is that my A18 wasn't capable of supplying the current needed to pump >> down >> >> the tube but the external power supply was. >> >>> Note: I proceeded very carefully with the external supply, letting it >> >> run for days at lower voltages and gradually working up to 3500. >> Also, I >> >> kept the current limit set as low as it would go. Anyway, it worked >> for >> >> me. Good luck with yours. >> >>> Rodger >> >>> >> >>> -----Original Message----- >> >>> From: Jim Muehlberg via time-nuts <time-nuts@lists.febo.com> >> >>> Sent: Wednesday, November 9, 2022 11:07 AM >> >>> To: time-nuts@lists.febo.com >> >>> Cc: Jim Muehlberg <jmuehlbe@nrao.edu> >> >>> Subject: [time-nuts] 5061A HV Supply, et al. >> >>> >> >>> I've recently come into possession of a 5061A. It is claimed that the >> >> tube is exhausted, but I'm unwilling to scrap it (horrors!) yet. I'm >> not >> >> well equipped for HV measurements, but it appears the +3500 V supply >> is not >> >> working. So, I'm hoping the tube just needs to be pumped since it >> hasn't >> >> been powered for years. Also missing the A3 assembly, but I'll worry >> about >> >> that later... >> >>> Anyone willing to part with their A18 or loan it to me? >> >>> >> >>> Thanks! >> >>> >> >> -- >> >> >> >> Jim Muehlberg >> >> Senior Engineer >> >> >> >> National Radio Astronomy Observatory >> >> ngVLA Local Oscillator Lead >> >> >> >> 1180 Boxwood Estates Rd B-111 >> >> Charlottesville, VA 22903-4602 >> >> P 434.296.0270 >> >> F 434.296.0324 >> >> www.cv.nrao.edu/~jmuehlbe >> >> _______________________________________________ >> >> time-nuts mailing list -- time-nuts@lists.febo.com >> >> To unsubscribe send an email to time-nuts-leave@lists.febo.com >> > _______________________________________________ >> > time-nuts mailing list -- time-nuts@lists.febo.com >> > To unsubscribe send an email to time-nuts-leave@lists.febo.com >> -- >> >> Jim Muehlberg >> Senior Engineer >> >> National Radio Astronomy Observatory >> ngVLA Local Oscillator Lead >> >> 1180 Boxwood Estates Rd B-111 >> Charlottesville, VA 22903-4602 >> P 434.296.0270 >> F 434.296.0324 >> www.cv.nrao.edu/~jmuehlbe >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe send an email to time-nuts-leave@lists.febo.com > >
PS
paul swed
Tue, Nov 15, 2022 1:37 PM

In general purpose transformers 60 Hz is fine but the HV transformer is KHz
as a heads up.
Paul
WB8TSL

On Mon, Nov 14, 2022 at 9:43 PM Lester Veenstra via time-nuts <
time-nuts@lists.febo.com> wrote:

Of course  a small “Variac” on the primary will get the voltage to what
you want

Lester B Veenstra  K1YCM  MØYCM  W8YCM  6Y6Y W8YCM/6Y 6Y8LV (Reformed
USNSG CTM1)

mailto:lester@veenstras.com lester@veenstras.com

452 Stable Ln

Keyser WV 26726 USA

GPS: 39.336826 N  78.982287 W (Google)

GPS: 39.33682 N  78.9823741 W (GPSDO)

Telephones:

Home:            +1-304-289-6057

US cell          +1-304-790-9192

Jamaica cell:    +1-876-456-8898

From: Andy Talbot [mailto:andy.g4jnt@gmail.com]
Sent: Monday, November 14, 2022 3:44 PM
To: Lester Veenstra
Cc: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Re: 5061A HV Supply, et al. {External}

I've got three transformers going spare ... :-)

Andy  G4JNT

IO90IV58AK12 (and others)

On Mon, 14 Nov 2022 at 20:40, Lester Veenstra m0ycm@veenstras.com wrote:

Sacrifice an old microwave oven to the Time Gods for its transformer

GPS: 39.336826 N  78.982287 W (Google)  =                    FM09MI20DU01
GPS: 39.33682 N  78.9823741 W (GPSDO) =                    FM09MI20CU70

Telephones:
Home:            +1-304-289-6057
US cell          +1-304-790-9192
Jamaica cell:    +1-876-456-8898

-----Original Message-----
From: Andy Talbot via time-nuts [mailto:time-nuts@lists.febo.com]
Sent: Monday, November 14, 2022 3:06 PM
To: Discussion of precise time and frequency measurement
Cc: Jim Muehlberg; Andy Talbot
Subject: [time-nuts] Re: 5061A HV Supply, et al. {External}

A simple flyback SMPSU with a Cockcroft-Walton tied on the end will give
you a simple high voltage supply.  A chain of high value resistors and you
can even stabilise it.

Andy
www.g4jnt.com

On Mon, 14 Nov 2022 at 20:00, Jim Muehlberg via time-nuts <
time-nuts@lists.febo.com> wrote:

Rodger,

Thanks for the encouragement!  I'm trying not to spend too much effort
on this - sort of a distraction from real work.  One would think that
NRAO would have a HV power supply lying around somewhere, but astronomy
has gone solid state now for decades!  We have a grad student trying to
nurse along another 5061 and we'd like to have a pair. I guess I'll
slice open the A18 module and see what makes it tick.

Thanks again,

Jim

On 11/10/2022 9:11 AM, Rodger via time-nuts wrote:

Hey Jim,

I got a 5061a a couple years ago with high ION pump current.  In my

case

the 3500 volts was there but the tube was not pumping down, even after
being on for weeks.  I found a Terranova 741 Ion pump power supply on

eBay

for under $100.  It was listed as not working but worked AOK when I got
it.  It's a nice supply with variable output voltage, current limiting,
safety interlocks, etc.  I hooked it up to the tube and let it run for a
few weeks.  Tube pumped down successfully and is working AOK now.  My

guess

is that my A18 wasn't capable of supplying the current needed to pump

down

the tube but the external power supply was.

Note: I proceeded very carefully with the external supply, letting it

run for days at lower voltages and gradually working up to 3500.  Also, I
kept the current limit set as low as it would go.  Anyway, it worked for
me.  Good luck with yours.

Rodger

-----Original Message-----
From: Jim Muehlberg via time-nuts time-nuts@lists.febo.com
Sent: Wednesday, November 9, 2022 11:07 AM
To: time-nuts@lists.febo.com
Cc: Jim Muehlberg jmuehlbe@nrao.edu
Subject: [time-nuts] 5061A HV Supply, et al.

I've recently come into possession of a 5061A.  It is claimed that the

tube is exhausted, but I'm unwilling to scrap it (horrors!) yet.  I'm not
well equipped for HV measurements, but it appears the +3500 V supply is

not

working.  So, I'm hoping the tube just needs to be pumped since it hasn't
been powered for years.  Also missing the A3 assembly, but I'll worry

about

that later...

Anyone willing to part with their A18 or loan it to me?

Thanks!

--

Jim Muehlberg
Senior Engineer

National Radio Astronomy Observatory
ngVLA Local Oscillator Lead

1180 Boxwood Estates Rd B-111
Charlottesville, VA 22903-4602
P 434.296.0270
F 434.296.0324
www.cv.nrao.edu/~jmuehlbe


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

In general purpose transformers 60 Hz is fine but the HV transformer is KHz as a heads up. Paul WB8TSL On Mon, Nov 14, 2022 at 9:43 PM Lester Veenstra via time-nuts < time-nuts@lists.febo.com> wrote: > Of course a small “Variac” on the primary will get the voltage to what > you want > > > > Lester B Veenstra K1YCM MØYCM W8YCM 6Y6Y W8YCM/6Y 6Y8LV (Reformed > USNSG CTM1) > > <mailto:lester@veenstras.com> lester@veenstras.com > > > > 452 Stable Ln > > Keyser WV 26726 USA > > > > GPS: 39.336826 N 78.982287 W (Google) > > GPS: 39.33682 N 78.9823741 W (GPSDO) > > > > > > Telephones: > > Home: +1-304-289-6057 > > US cell +1-304-790-9192 > > Jamaica cell: +1-876-456-8898 > > > > From: Andy Talbot [mailto:andy.g4jnt@gmail.com] > Sent: Monday, November 14, 2022 3:44 PM > To: Lester Veenstra > Cc: Discussion of precise time and frequency measurement > Subject: Re: [time-nuts] Re: 5061A HV Supply, et al. {External} > > > > I've got three transformers going spare ... :-) > > > > Andy G4JNT > > IO90IV58AK12 (and others) > > > > > > > > > > On Mon, 14 Nov 2022 at 20:40, Lester Veenstra <m0ycm@veenstras.com> wrote: > > Sacrifice an old microwave oven to the Time Gods for its transformer > > GPS: 39.336826 N 78.982287 W (Google) = FM09MI20DU01 > GPS: 39.33682 N 78.9823741 W (GPSDO) = FM09MI20CU70 > > > > > > > Telephones: > Home: +1-304-289-6057 > US cell +1-304-790-9192 > Jamaica cell: +1-876-456-8898 > > -----Original Message----- > From: Andy Talbot via time-nuts [mailto:time-nuts@lists.febo.com] > Sent: Monday, November 14, 2022 3:06 PM > To: Discussion of precise time and frequency measurement > Cc: Jim Muehlberg; Andy Talbot > Subject: [time-nuts] Re: 5061A HV Supply, et al. {External} > > A simple flyback SMPSU with a Cockcroft-Walton tied on the end will give > you a simple high voltage supply. A chain of high value resistors and you > can even stabilise it. > > Andy > www.g4jnt.com > > > > On Mon, 14 Nov 2022 at 20:00, Jim Muehlberg via time-nuts < > time-nuts@lists.febo.com> wrote: > > > Rodger, > > > > Thanks for the encouragement! I'm trying not to spend too much effort > > on this - sort of a distraction from real work. One would think that > > NRAO would have a HV power supply lying around somewhere, but astronomy > > has gone solid state now for decades! We have a grad student trying to > > nurse along another 5061 and we'd like to have a pair. I guess I'll > > slice open the A18 module and see what makes it tick. > > > > Thanks again, > > > > Jim > > > > On 11/10/2022 9:11 AM, Rodger via time-nuts wrote: > > > Hey Jim, > > > > > > I got a 5061a a couple years ago with high ION pump current. In my > case > > the 3500 volts was there but the tube was not pumping down, even after > > being on for weeks. I found a Terranova 741 Ion pump power supply on > eBay > > for under $100. It was listed as not working but worked AOK when I got > > it. It's a nice supply with variable output voltage, current limiting, > > safety interlocks, etc. I hooked it up to the tube and let it run for a > > few weeks. Tube pumped down successfully and is working AOK now. My > guess > > is that my A18 wasn't capable of supplying the current needed to pump > down > > the tube but the external power supply was. > > > > > > Note: I proceeded very carefully with the external supply, letting it > > run for days at lower voltages and gradually working up to 3500. Also, I > > kept the current limit set as low as it would go. Anyway, it worked for > > me. Good luck with yours. > > > > > > Rodger > > > > > > -----Original Message----- > > > From: Jim Muehlberg via time-nuts <time-nuts@lists.febo.com> > > > Sent: Wednesday, November 9, 2022 11:07 AM > > > To: time-nuts@lists.febo.com > > > Cc: Jim Muehlberg <jmuehlbe@nrao.edu> > > > Subject: [time-nuts] 5061A HV Supply, et al. > > > > > > I've recently come into possession of a 5061A. It is claimed that the > > tube is exhausted, but I'm unwilling to scrap it (horrors!) yet. I'm not > > well equipped for HV measurements, but it appears the +3500 V supply is > not > > working. So, I'm hoping the tube just needs to be pumped since it hasn't > > been powered for years. Also missing the A3 assembly, but I'll worry > about > > that later... > > > > > > Anyone willing to part with their A18 or loan it to me? > > > > > > Thanks! > > > > > -- > > > > Jim Muehlberg > > Senior Engineer > > > > National Radio Astronomy Observatory > > ngVLA Local Oscillator Lead > > > > 1180 Boxwood Estates Rd B-111 > > Charlottesville, VA 22903-4602 > > P 434.296.0270 > > F 434.296.0324 > > www.cv.nrao.edu/~jmuehlbe > > _______________________________________________ > > time-nuts mailing list -- time-nuts@lists.febo.com > > To unsubscribe send an email to time-nuts-leave@lists.febo.com > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com
LV
Lester Veenstra
Tue, Nov 15, 2022 2:46 PM

Not in my ovans.  They run directly oiff od 50/60 Hz line.  There is no RF freq flyback involved.

B>T>W>  These would be useful with such a variac & transformer test supply.

https://www.ebay.com/vod/FetchOrderDetails?itemId=402568223370&transactionId=1108957800027&mkevt=1&mkpid=0&emsid=e11401.m1842.l5920&mkcid=7&ch=osgood&euid=78ef2d648c414a41a24157fcd09254de&bu=43198945816&exe=102809&ext=241959&osub=-1%7E1&crd=20221115061028&segname=11401 US Stock 4pcs CL01-12 Microwave Oven High Voltage Diode Rectifier

https://www.ebay.com/vod/FetchOrderDetails?itemId=402568223370&transactionId=1108957800027&mkevt=1&mkpid=0&emsid=e11401.m1842.l5919&mkcid=7&ch=osgood&euid=78ef2d648c414a41a24157fcd09254de&bu=43198945816&exe=102809&ext=241959&osub=-1%7E1&crd=20221115061028&segname=11401 US Stock 4pcs CL01-12 Microwave Oven High Voltage Diode Rectifier

Total: $7.47

Item ID: 402568223370

Lester B Veenstra  K1YCM  MØYCM  W8YCM  6Y6Y W8YCM/6Y 6Y8LV (Reformed USNSG CTM1)

mailto:lester@veenstras.com lester@veenstras.com

452 Stable Ln

Keyser WV 26726 USA

GPS: 39.336826 N  78.982287 W (Google)

GPS: 39.33682 N  78.9823741 W (GPSDO)

Telephones:

Home:            +1-304-289-6057

US cell          +1-304-790-9192

Jamaica cell:    +1-876-456-8898

From: paul swed [mailto:paulswedb@gmail.com]
Sent: Tuesday, November 15, 2022 8:37 AM
To: Discussion of precise time and frequency measurement
Cc: Andy Talbot; Lester Veenstra
Subject: Re: [time-nuts] Re: 5061A HV Supply, et al. {External}

In general purpose transformers 60 Hz is fine but the HV transformer is KHz as a heads up.

Paul

WB8TSL

On Mon, Nov 14, 2022 at 9:43 PM Lester Veenstra via time-nuts time-nuts@lists.febo.com wrote:

Of course  a small “Variac” on the primary will get the voltage to what you want

Lester B Veenstra  K1YCM  MØYCM  W8YCM  6Y6Y W8YCM/6Y 6Y8LV (Reformed USNSG CTM1)

mailto:lester@veenstras.com lester@veenstras.com

452 Stable Ln

Keyser WV 26726 USA

GPS: 39.336826 N  78.982287 W (Google)

GPS: 39.33682 N  78.9823741 W (GPSDO)

Telephones:

Home:            +1-304-289-6057

US cell          +1-304-790-9192

Jamaica cell:    +1-876-456-8898

From: Andy Talbot [mailto:andy.g4jnt@gmail.com]
Sent: Monday, November 14, 2022 3:44 PM
To: Lester Veenstra
Cc: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Re: 5061A HV Supply, et al. {External}

I've got three transformers going spare ... :-)

Andy  G4JNT

IO90IV58AK12 (and others)

On Mon, 14 Nov 2022 at 20:40, Lester Veenstra m0ycm@veenstras.com wrote:

Sacrifice an old microwave oven to the Time Gods for its transformer

GPS: 39.336826 N  78.982287 W (Google)  =                    FM09MI20DU01
GPS: 39.33682 N  78.9823741 W (GPSDO) =                    FM09MI20CU70

Telephones:
Home:            +1-304-289-6057
US cell          +1-304-790-9192
Jamaica cell:    +1-876-456-8898

-----Original Message-----
From: Andy Talbot via time-nuts [mailto:time-nuts@lists.febo.com]
Sent: Monday, November 14, 2022 3:06 PM
To: Discussion of precise time and frequency measurement
Cc: Jim Muehlberg; Andy Talbot
Subject: [time-nuts] Re: 5061A HV Supply, et al. {External}

A simple flyback SMPSU with a Cockcroft-Walton tied on the end will give
you a simple high voltage supply.  A chain of high value resistors and you
can even stabilise it.

Andy
www.g4jnt.com

On Mon, 14 Nov 2022 at 20:00, Jim Muehlberg via time-nuts <
time-nuts@lists.febo.com> wrote:

Rodger,

Thanks for the encouragement!  I'm trying not to spend too much effort
on this - sort of a distraction from real work.  One would think that
NRAO would have a HV power supply lying around somewhere, but astronomy
has gone solid state now for decades!  We have a grad student trying to
nurse along another 5061 and we'd like to have a pair. I guess I'll
slice open the A18 module and see what makes it tick.

Thanks again,

Jim

On 11/10/2022 9:11 AM, Rodger via time-nuts wrote:

Hey Jim,

I got a 5061a a couple years ago with high ION pump current.  In my case

the 3500 volts was there but the tube was not pumping down, even after
being on for weeks.  I found a Terranova 741 Ion pump power supply on eBay
for under $100.  It was listed as not working but worked AOK when I got
it.  It's a nice supply with variable output voltage, current limiting,
safety interlocks, etc.  I hooked it up to the tube and let it run for a
few weeks.  Tube pumped down successfully and is working AOK now.  My

guess

is that my A18 wasn't capable of supplying the current needed to pump down
the tube but the external power supply was.

Note: I proceeded very carefully with the external supply, letting it

run for days at lower voltages and gradually working up to 3500.  Also, I
kept the current limit set as low as it would go.  Anyway, it worked for
me.  Good luck with yours.

Rodger

-----Original Message-----
From: Jim Muehlberg via time-nuts time-nuts@lists.febo.com
Sent: Wednesday, November 9, 2022 11:07 AM
To: time-nuts@lists.febo.com
Cc: Jim Muehlberg jmuehlbe@nrao.edu
Subject: [time-nuts] 5061A HV Supply, et al.

I've recently come into possession of a 5061A.  It is claimed that the

tube is exhausted, but I'm unwilling to scrap it (horrors!) yet.  I'm not
well equipped for HV measurements, but it appears the +3500 V supply is

not

working.  So, I'm hoping the tube just needs to be pumped since it hasn't
been powered for years.  Also missing the A3 assembly, but I'll worry

about

that later...

Anyone willing to part with their A18 or loan it to me?

Thanks!

--

Jim Muehlberg
Senior Engineer

National Radio Astronomy Observatory
ngVLA Local Oscillator Lead

1180 Boxwood Estates Rd B-111
Charlottesville, VA 22903-4602
P 434.296.0270
F 434.296.0324
www.cv.nrao.edu/~jmuehlbe


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

Not in my ovans. They run directly oiff od 50/60 Hz line. There is no RF freq flyback involved. B>T>W> These would be useful with such a variac & transformer test supply. <https://www.ebay.com/vod/FetchOrderDetails?itemId=402568223370&transactionId=1108957800027&mkevt=1&mkpid=0&emsid=e11401.m1842.l5920&mkcid=7&ch=osgood&euid=78ef2d648c414a41a24157fcd09254de&bu=43198945816&exe=102809&ext=241959&osub=-1%7E1&crd=20221115061028&segname=11401> US Stock 4pcs CL01-12 Microwave Oven High Voltage Diode Rectifier <https://www.ebay.com/vod/FetchOrderDetails?itemId=402568223370&transactionId=1108957800027&mkevt=1&mkpid=0&emsid=e11401.m1842.l5919&mkcid=7&ch=osgood&euid=78ef2d648c414a41a24157fcd09254de&bu=43198945816&exe=102809&ext=241959&osub=-1%7E1&crd=20221115061028&segname=11401> US Stock 4pcs CL01-12 Microwave Oven High Voltage Diode Rectifier Total: $7.47 Item ID: 402568223370 Lester B Veenstra K1YCM MØYCM W8YCM 6Y6Y W8YCM/6Y 6Y8LV (Reformed USNSG CTM1) <mailto:lester@veenstras.com> lester@veenstras.com 452 Stable Ln Keyser WV 26726 USA GPS: 39.336826 N 78.982287 W (Google) GPS: 39.33682 N 78.9823741 W (GPSDO) Telephones: Home: +1-304-289-6057 US cell +1-304-790-9192 Jamaica cell: +1-876-456-8898 From: paul swed [mailto:paulswedb@gmail.com] Sent: Tuesday, November 15, 2022 8:37 AM To: Discussion of precise time and frequency measurement Cc: Andy Talbot; Lester Veenstra Subject: Re: [time-nuts] Re: 5061A HV Supply, et al. {External} In general purpose transformers 60 Hz is fine but the HV transformer is KHz as a heads up. Paul WB8TSL On Mon, Nov 14, 2022 at 9:43 PM Lester Veenstra via time-nuts <time-nuts@lists.febo.com> wrote: Of course a small “Variac” on the primary will get the voltage to what you want Lester B Veenstra K1YCM MØYCM W8YCM 6Y6Y W8YCM/6Y 6Y8LV (Reformed USNSG CTM1) <mailto:lester@veenstras.com> lester@veenstras.com 452 Stable Ln Keyser WV 26726 USA GPS: 39.336826 N 78.982287 W (Google) GPS: 39.33682 N 78.9823741 W (GPSDO) Telephones: Home: +1-304-289-6057 US cell +1-304-790-9192 Jamaica cell: +1-876-456-8898 From: Andy Talbot [mailto:andy.g4jnt@gmail.com] Sent: Monday, November 14, 2022 3:44 PM To: Lester Veenstra Cc: Discussion of precise time and frequency measurement Subject: Re: [time-nuts] Re: 5061A HV Supply, et al. {External} I've got three transformers going spare ... :-) Andy G4JNT IO90IV58AK12 (and others) On Mon, 14 Nov 2022 at 20:40, Lester Veenstra <m0ycm@veenstras.com> wrote: Sacrifice an old microwave oven to the Time Gods for its transformer GPS: 39.336826 N 78.982287 W (Google) = FM09MI20DU01 GPS: 39.33682 N 78.9823741 W (GPSDO) = FM09MI20CU70 Telephones: Home: +1-304-289-6057 US cell +1-304-790-9192 Jamaica cell: +1-876-456-8898 -----Original Message----- From: Andy Talbot via time-nuts [mailto:time-nuts@lists.febo.com] Sent: Monday, November 14, 2022 3:06 PM To: Discussion of precise time and frequency measurement Cc: Jim Muehlberg; Andy Talbot Subject: [time-nuts] Re: 5061A HV Supply, et al. {External} A simple flyback SMPSU with a Cockcroft-Walton tied on the end will give you a simple high voltage supply. A chain of high value resistors and you can even stabilise it. Andy www.g4jnt.com On Mon, 14 Nov 2022 at 20:00, Jim Muehlberg via time-nuts < time-nuts@lists.febo.com> wrote: > Rodger, > > Thanks for the encouragement! I'm trying not to spend too much effort > on this - sort of a distraction from real work. One would think that > NRAO would have a HV power supply lying around somewhere, but astronomy > has gone solid state now for decades! We have a grad student trying to > nurse along another 5061 and we'd like to have a pair. I guess I'll > slice open the A18 module and see what makes it tick. > > Thanks again, > > Jim > > On 11/10/2022 9:11 AM, Rodger via time-nuts wrote: > > Hey Jim, > > > > I got a 5061a a couple years ago with high ION pump current. In my case > the 3500 volts was there but the tube was not pumping down, even after > being on for weeks. I found a Terranova 741 Ion pump power supply on eBay > for under $100. It was listed as not working but worked AOK when I got > it. It's a nice supply with variable output voltage, current limiting, > safety interlocks, etc. I hooked it up to the tube and let it run for a > few weeks. Tube pumped down successfully and is working AOK now. My guess > is that my A18 wasn't capable of supplying the current needed to pump down > the tube but the external power supply was. > > > > Note: I proceeded very carefully with the external supply, letting it > run for days at lower voltages and gradually working up to 3500. Also, I > kept the current limit set as low as it would go. Anyway, it worked for > me. Good luck with yours. > > > > Rodger > > > > -----Original Message----- > > From: Jim Muehlberg via time-nuts <time-nuts@lists.febo.com> > > Sent: Wednesday, November 9, 2022 11:07 AM > > To: time-nuts@lists.febo.com > > Cc: Jim Muehlberg <jmuehlbe@nrao.edu> > > Subject: [time-nuts] 5061A HV Supply, et al. > > > > I've recently come into possession of a 5061A. It is claimed that the > tube is exhausted, but I'm unwilling to scrap it (horrors!) yet. I'm not > well equipped for HV measurements, but it appears the +3500 V supply is not > working. So, I'm hoping the tube just needs to be pumped since it hasn't > been powered for years. Also missing the A3 assembly, but I'll worry about > that later... > > > > Anyone willing to part with their A18 or loan it to me? > > > > Thanks! > > > -- > > Jim Muehlberg > Senior Engineer > > National Radio Astronomy Observatory > ngVLA Local Oscillator Lead > > 1180 Boxwood Estates Rd B-111 > Charlottesville, VA 22903-4602 > P 434.296.0270 > F 434.296.0324 > www.cv.nrao.edu/~jmuehlbe > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com _______________________________________________ time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe send an email to time-nuts-leave@lists.febo.com _______________________________________________ time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe send an email to time-nuts-leave@lists.febo.com