Sun, Oct 14, 2012 2:18 AM
Dear Group,
Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
- Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
- If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
- There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
Kindest regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
Dear Group,
Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
1. Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
2. If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
3. There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
Kindest regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
AB
Azelio Boriani
Sun, Oct 14, 2012 12:47 PM
Use a multimeter to see if the DB25 pin 2 is -12V, if so you have a
standard RS232 serial port. Then connect to it, use a USB-RS232 interface
if necessary, and type SYST:STAT?
you will see the status page out of the Z3805 (use 9600,N,8,1 for the
serial settings).
On Sun, Oct 14, 2012 at 4:18 AM, Edgardo Molina xe1xus@amsat.org wrote:
Dear Group,
Good evening. I bought two HP Z3805A units a few weeks ago. They were bare
units without antenna or power supply. After struggling to find a suitable
24V 3A noise free power supply for each of them, I mounted a couple of
Symmetricom antennas and just fired them up.
After around 25 minutes, the GPS Lock light lighted up. As the manual for
this particular model is unobtainable in several searches through the
Internet, I found that the readily available manual for the Z3801A would
cover much of the Z3905A operation. I also got a software for monitoring
the operation from BD Systems Inc. In Colorado. Now that I am sure both are
operating and locked only by looking at the front panel lights, I would
like to ask a couple of questions regarding their communication to the PC.
- Is the DB-25 serial port carrying RS-232 signals? I see the port of the
Z380A1 is different: RS-422.
- If the port is plain RS-232 is it possible to use a Serial to USB
adapter? I have heard about the inconsistencies that such an adapter could
cause when used with a TBolt. Should I assume this is the same case with
the Z3805A?
- There was once a project for building a TBolt state monitor on a 2 line
LCD screen. I own three of those monitors and they are just great. Is
there an equivalent for these HP GPS units?
Any suggestions for the operation of these units and your ever expert and
kind advise is always welcome. You all have a nice weekend!
Kindest regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario
de este mensaje, le suplicamos se lo notifique al remitente mediante un
correo electrónico y que borre el presente mensaje y sus anexos de su
computadora sin retener una copia de los mismos. Queda estrictamente
prohibido copiar este mensaje o hacer usode el para cualquier propósito o
divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are
not the intended recipient please immediately advise the sender by replying
to this e-mail and then deleting the message and its attachments from your
computer without keeping a copy. It is strictly forbidden to copy it or use
it for any purpose or disclose its contents to any third party. Thank you.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Use a multimeter to see if the DB25 pin 2 is -12V, if so you have a
standard RS232 serial port. Then connect to it, use a USB-RS232 interface
if necessary, and type SYST:STAT?
you will see the status page out of the Z3805 (use 9600,N,8,1 for the
serial settings).
On Sun, Oct 14, 2012 at 4:18 AM, Edgardo Molina <xe1xus@amsat.org> wrote:
> Dear Group,
>
> Good evening. I bought two HP Z3805A units a few weeks ago. They were bare
> units without antenna or power supply. After struggling to find a suitable
> 24V 3A noise free power supply for each of them, I mounted a couple of
> Symmetricom antennas and just fired them up.
>
> After around 25 minutes, the GPS Lock light lighted up. As the manual for
> this particular model is unobtainable in several searches through the
> Internet, I found that the readily available manual for the Z3801A would
> cover much of the Z3905A operation. I also got a software for monitoring
> the operation from BD Systems Inc. In Colorado. Now that I am sure both are
> operating and locked only by looking at the front panel lights, I would
> like to ask a couple of questions regarding their communication to the PC.
>
> 1. Is the DB-25 serial port carrying RS-232 signals? I see the port of the
> Z380A1 is different: RS-422.
> 2. If the port is plain RS-232 is it possible to use a Serial to USB
> adapter? I have heard about the inconsistencies that such an adapter could
> cause when used with a TBolt. Should I assume this is the same case with
> the Z3805A?
> 3. There was once a project for building a TBolt state monitor on a 2 line
> LCD screen. I own three of those monitors and they are just great. Is
> there an equivalent for these HP GPS units?
>
> Any suggestions for the operation of these units and your ever expert and
> kind advise is always welcome. You all have a nice weekend!
>
> Kindest regards,
>
>
>
> Edgardo Molina
> Dirección IPTEL
>
> www.iptel.net.mx
>
> T : 55 55 55202444
> M : 04455 20501854
>
> Piensa en Bits SA de CV
>
>
>
> Información anexa:
>
>
>
>
> CONFIDENCIALIDAD DE INFORMACION
>
> Este mensaje tiene carácter confidencial. Si usted no es el destinarario
> de este mensaje, le suplicamos se lo notifique al remitente mediante un
> correo electrónico y que borre el presente mensaje y sus anexos de su
> computadora sin retener una copia de los mismos. Queda estrictamente
> prohibido copiar este mensaje o hacer usode el para cualquier propósito o
> divulgar su en forma parcial o total su contenido. Gracias.
>
>
> NON-DISCLOSURE OF INFORMATION
>
> This email is strictly confidential and may also be privileged. If you are
> not the intended recipient please immediately advise the sender by replying
> to this e-mail and then deleting the message and its attachments from your
> computer without keeping a copy. It is strictly forbidden to copy it or use
> it for any purpose or disclose its contents to any third party. Thank you.
>
>
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
JA
John Ackermann N8UR
Sun, Oct 14, 2012 1:23 PM
On the unit I bought from fluke.l, with a "Samsung" label, there are two
serial ports. The bottom one is the control port and is definitely
RS-232. The top port has only three wires (TXD, RXD, SG) and I think is
used for some sort of monitoring.
I've been using Ulrich's "z38xx" software to monitor it with good success.
John
Edgardo Molina said the following on 10/13/2012 10:18 PM:
Dear Group,
Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
- Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
- If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
- There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
Kindest regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
On the unit I bought from fluke.l, with a "Samsung" label, there are two
serial ports. The bottom one is the control port and is definitely
RS-232. The top port has only three wires (TXD, RXD, SG) and I think is
used for some sort of monitoring.
I've been using Ulrich's "z38xx" software to monitor it with good success.
John
----
Edgardo Molina said the following on 10/13/2012 10:18 PM:
> Dear Group,
>
> Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
>
> After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
>
> 1. Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
> 2. If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
> 3. There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
>
> Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
>
> Kindest regards,
>
>
>
> Edgardo Molina
> Dirección IPTEL
>
> www.iptel.net.mx
>
> T : 55 55 55202444
> M : 04455 20501854
>
> Piensa en Bits SA de CV
>
>
>
> Información anexa:
>
>
>
>
> CONFIDENCIALIDAD DE INFORMACION
>
> Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
>
>
> NON-DISCLOSURE OF INFORMATION
>
> This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
>
>
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
BC
Bob Camp
Sun, Oct 14, 2012 1:39 PM
Hi
I've been using Ulrich's software on my 3805's as well.
Trimble and HP took different approaches on their GPSDO's. Trimble lets you get into things and fiddle around. HP is much more a black box. I have never seen anything that lets you "fiddle' with one of the HP units. Things like loop gain, time constant, DAC settings that are configurable on the Trimble are not accessible on the HP. HP only lets you do the satellite mask and re-survey sort of stuff. You can monitor status, you just can't do much about it...
Bob
On Oct 14, 2012, at 9:23 AM, John Ackermann N8UR jra@febo.com wrote:
On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
I've been using Ulrich's "z38xx" software to monitor it with good success.
John
Edgardo Molina said the following on 10/13/2012 10:18 PM:
Dear Group,
Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
- Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
- If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
- There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
Kindest regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Hi
I've been using Ulrich's software on my 3805's as well.
Trimble and HP took different approaches on their GPSDO's. Trimble lets you get into things and fiddle around. HP is much more a black box. I have never seen anything that lets you "fiddle' with one of the HP units. Things like loop gain, time constant, DAC settings that are configurable on the Trimble are not accessible on the HP. HP only lets you do the satellite mask and re-survey sort of stuff. You can monitor status, you just can't do much about it...
Bob
On Oct 14, 2012, at 9:23 AM, John Ackermann N8UR <jra@febo.com> wrote:
> On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
>
> I've been using Ulrich's "z38xx" software to monitor it with good success.
>
> John
> ----
>
> Edgardo Molina said the following on 10/13/2012 10:18 PM:
>> Dear Group,
>>
>> Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
>>
>> After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
>>
>> 1. Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
>> 2. If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
>> 3. There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
>>
>> Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
>>
>> Kindest regards,
>>
>>
>>
>> Edgardo Molina
>> Dirección IPTEL
>>
>> www.iptel.net.mx
>>
>> T : 55 55 55202444
>> M : 04455 20501854
>>
>> Piensa en Bits SA de CV
>>
>>
>>
>> Información anexa:
>>
>>
>>
>>
>> CONFIDENCIALIDAD DE INFORMACION
>>
>> Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
>>
>>
>> NON-DISCLOSURE OF INFORMATION
>>
>> This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
>>
>>
>>
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
EM
Edgardo Molina
Sun, Oct 14, 2012 4:24 PM
Dear Azelio,
Thank you. I left the receivers running since yesterday and they are still solid lock state. I will measure pin 2 as you kindly stated. I will come back with the results. Thank you for pointing out this test. You are always willing to help!
Kind regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 7:47 AM, Azelio Boriani azelio.boriani@screen.it wrote:
Use a multimeter to see if the DB25 pin 2 is -12V, if so you have a
standard RS232 serial port. Then connect to it, use a USB-RS232 interface
if necessary, and type SYST:STAT?
you will see the status page out of the Z3805 (use 9600,N,8,1 for the
serial settings).
On Sun, Oct 14, 2012 at 4:18 AM, Edgardo Molina xe1xus@amsat.org wrote:
Dear Group,
Good evening. I bought two HP Z3805A units a few weeks ago. They were bare
units without antenna or power supply. After struggling to find a suitable
24V 3A noise free power supply for each of them, I mounted a couple of
Symmetricom antennas and just fired them up.
After around 25 minutes, the GPS Lock light lighted up. As the manual for
this particular model is unobtainable in several searches through the
Internet, I found that the readily available manual for the Z3801A would
cover much of the Z3905A operation. I also got a software for monitoring
the operation from BD Systems Inc. In Colorado. Now that I am sure both are
operating and locked only by looking at the front panel lights, I would
like to ask a couple of questions regarding their communication to the PC.
- Is the DB-25 serial port carrying RS-232 signals? I see the port of the
Z380A1 is different: RS-422.
- If the port is plain RS-232 is it possible to use a Serial to USB
adapter? I have heard about the inconsistencies that such an adapter could
cause when used with a TBolt. Should I assume this is the same case with
the Z3805A?
- There was once a project for building a TBolt state monitor on a 2 line
LCD screen. I own three of those monitors and they are just great. Is
there an equivalent for these HP GPS units?
Any suggestions for the operation of these units and your ever expert and
kind advise is always welcome. You all have a nice weekend!
Kindest regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario
de este mensaje, le suplicamos se lo notifique al remitente mediante un
correo electrónico y que borre el presente mensaje y sus anexos de su
computadora sin retener una copia de los mismos. Queda estrictamente
prohibido copiar este mensaje o hacer usode el para cualquier propósito o
divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are
not the intended recipient please immediately advise the sender by replying
to this e-mail and then deleting the message and its attachments from your
computer without keeping a copy. It is strictly forbidden to copy it or use
it for any purpose or disclose its contents to any third party. Thank you.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Dear Azelio,
Thank you. I left the receivers running since yesterday and they are still solid lock state. I will measure pin 2 as you kindly stated. I will come back with the results. Thank you for pointing out this test. You are always willing to help!
Kind regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 7:47 AM, Azelio Boriani <azelio.boriani@screen.it> wrote:
> Use a multimeter to see if the DB25 pin 2 is -12V, if so you have a
> standard RS232 serial port. Then connect to it, use a USB-RS232 interface
> if necessary, and type SYST:STAT?
> you will see the status page out of the Z3805 (use 9600,N,8,1 for the
> serial settings).
>
> On Sun, Oct 14, 2012 at 4:18 AM, Edgardo Molina <xe1xus@amsat.org> wrote:
>
>> Dear Group,
>>
>> Good evening. I bought two HP Z3805A units a few weeks ago. They were bare
>> units without antenna or power supply. After struggling to find a suitable
>> 24V 3A noise free power supply for each of them, I mounted a couple of
>> Symmetricom antennas and just fired them up.
>>
>> After around 25 minutes, the GPS Lock light lighted up. As the manual for
>> this particular model is unobtainable in several searches through the
>> Internet, I found that the readily available manual for the Z3801A would
>> cover much of the Z3905A operation. I also got a software for monitoring
>> the operation from BD Systems Inc. In Colorado. Now that I am sure both are
>> operating and locked only by looking at the front panel lights, I would
>> like to ask a couple of questions regarding their communication to the PC.
>>
>> 1. Is the DB-25 serial port carrying RS-232 signals? I see the port of the
>> Z380A1 is different: RS-422.
>> 2. If the port is plain RS-232 is it possible to use a Serial to USB
>> adapter? I have heard about the inconsistencies that such an adapter could
>> cause when used with a TBolt. Should I assume this is the same case with
>> the Z3805A?
>> 3. There was once a project for building a TBolt state monitor on a 2 line
>> LCD screen. I own three of those monitors and they are just great. Is
>> there an equivalent for these HP GPS units?
>>
>> Any suggestions for the operation of these units and your ever expert and
>> kind advise is always welcome. You all have a nice weekend!
>>
>> Kindest regards,
>>
>>
>>
>> Edgardo Molina
>> Dirección IPTEL
>>
>> www.iptel.net.mx
>>
>> T : 55 55 55202444
>> M : 04455 20501854
>>
>> Piensa en Bits SA de CV
>>
>>
>>
>> Información anexa:
>>
>>
>>
>>
>> CONFIDENCIALIDAD DE INFORMACION
>>
>> Este mensaje tiene carácter confidencial. Si usted no es el destinarario
>> de este mensaje, le suplicamos se lo notifique al remitente mediante un
>> correo electrónico y que borre el presente mensaje y sus anexos de su
>> computadora sin retener una copia de los mismos. Queda estrictamente
>> prohibido copiar este mensaje o hacer usode el para cualquier propósito o
>> divulgar su en forma parcial o total su contenido. Gracias.
>>
>>
>> NON-DISCLOSURE OF INFORMATION
>>
>> This email is strictly confidential and may also be privileged. If you are
>> not the intended recipient please immediately advise the sender by replying
>> to this e-mail and then deleting the message and its attachments from your
>> computer without keeping a copy. It is strictly forbidden to copy it or use
>> it for any purpose or disclose its contents to any third party. Thank you.
>>
>>
>>
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
EM
Edgardo Molina
Sun, Oct 14, 2012 4:29 PM
Dear John,
Very interesting. I haven't had the time to compare the Samsung branded units with the typical HP units. Mine is the later. It only has one communications port, plus 2x10 MHz and 2x 1PPS. As soon as I am done with the late family breakfast I will do some testing and let you know what I find.
How does the oscillators from the Samsung units compare to the HP units? Are the Samsung units superior in any sense? Both have the same GPS receiver/number of channels?
Thank you for opening my curiosity regarding this GPSDO model.
Best regards!
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 8:23 AM, John Ackermann N8UR jra@febo.com wrote:
On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
I've been using Ulrich's "z38xx" software to monitor it with good success.
John
Edgardo Molina said the following on 10/13/2012 10:18 PM:
Dear Group,
Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
- Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
- If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
- There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
Kindest regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Dear John,
Very interesting. I haven't had the time to compare the Samsung branded units with the typical HP units. Mine is the later. It only has one communications port, plus 2x10 MHz and 2x 1PPS. As soon as I am done with the late family breakfast I will do some testing and let you know what I find.
How does the oscillators from the Samsung units compare to the HP units? Are the Samsung units superior in any sense? Both have the same GPS receiver/number of channels?
Thank you for opening my curiosity regarding this GPSDO model.
Best regards!
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 8:23 AM, John Ackermann N8UR <jra@febo.com> wrote:
> On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
>
> I've been using Ulrich's "z38xx" software to monitor it with good success.
>
> John
> ----
>
> Edgardo Molina said the following on 10/13/2012 10:18 PM:
>> Dear Group,
>>
>> Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
>>
>> After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
>>
>> 1. Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
>> 2. If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
>> 3. There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
>>
>> Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
>>
>> Kindest regards,
>>
>>
>>
>> Edgardo Molina
>> Dirección IPTEL
>>
>> www.iptel.net.mx
>>
>> T : 55 55 55202444
>> M : 04455 20501854
>>
>> Piensa en Bits SA de CV
>>
>>
>>
>> Información anexa:
>>
>>
>>
>>
>> CONFIDENCIALIDAD DE INFORMACION
>>
>> Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
>>
>>
>> NON-DISCLOSURE OF INFORMATION
>>
>> This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
>>
>>
>>
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
EM
Edgardo Molina
Sun, Oct 14, 2012 4:33 PM
Dear Bob,
I agree. I have been playing around with two TBolts for my thesis work and they are just great. HP as some circles have stated, are plug and forget GPSDOs. Still I love the looks, construction and the fact that they match with the rest of the HP equipment here in the lab.
As I continue working on both the HP and the TBolt units, I will try to share my experiences, despite they will not be up to the level some folks here have achieved. I am moving in small steps but gaining experience and learning in the process. Always with a good attitude and looking to have some fun during the process.
Thank you!
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 8:39 AM, Bob Camp lists@rtty.us wrote:
Hi
I've been using Ulrich's software on my 3805's as well.
Trimble and HP took different approaches on their GPSDO's. Trimble lets you get into things and fiddle around. HP is much more a black box. I have never seen anything that lets you "fiddle' with one of the HP units. Things like loop gain, time constant, DAC settings that are configurable on the Trimble are not accessible on the HP. HP only lets you do the satellite mask and re-survey sort of stuff. You can monitor status, you just can't do much about it...
Bob
On Oct 14, 2012, at 9:23 AM, John Ackermann N8UR jra@febo.com wrote:
On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
I've been using Ulrich's "z38xx" software to monitor it with good success.
John
Edgardo Molina said the following on 10/13/2012 10:18 PM:
Dear Group,
Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
- Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
- If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
- There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
Kindest regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Dear Bob,
I agree. I have been playing around with two TBolts for my thesis work and they are just great. HP as some circles have stated, are plug and forget GPSDOs. Still I love the looks, construction and the fact that they match with the rest of the HP equipment here in the lab.
As I continue working on both the HP and the TBolt units, I will try to share my experiences, despite they will not be up to the level some folks here have achieved. I am moving in small steps but gaining experience and learning in the process. Always with a good attitude and looking to have some fun during the process.
Thank you!
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 8:39 AM, Bob Camp <lists@rtty.us> wrote:
> Hi
>
> I've been using Ulrich's software on my 3805's as well.
>
> Trimble and HP took different approaches on their GPSDO's. Trimble lets you get into things and fiddle around. HP is much more a black box. I have never seen anything that lets you "fiddle' with one of the HP units. Things like loop gain, time constant, DAC settings that are configurable on the Trimble are not accessible on the HP. HP only lets you do the satellite mask and re-survey sort of stuff. You can monitor status, you just can't do much about it...
>
> Bob
>
> On Oct 14, 2012, at 9:23 AM, John Ackermann N8UR <jra@febo.com> wrote:
>
>> On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
>>
>> I've been using Ulrich's "z38xx" software to monitor it with good success.
>>
>> John
>> ----
>>
>> Edgardo Molina said the following on 10/13/2012 10:18 PM:
>>> Dear Group,
>>>
>>> Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
>>>
>>> After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
>>>
>>> 1. Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
>>> 2. If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
>>> 3. There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
>>>
>>> Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
>>>
>>> Kindest regards,
>>>
>>>
>>>
>>> Edgardo Molina
>>> Dirección IPTEL
>>>
>>> www.iptel.net.mx
>>>
>>> T : 55 55 55202444
>>> M : 04455 20501854
>>>
>>> Piensa en Bits SA de CV
>>>
>>>
>>>
>>> Información anexa:
>>>
>>>
>>>
>>>
>>> CONFIDENCIALIDAD DE INFORMACION
>>>
>>> Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
>>>
>>>
>>> NON-DISCLOSURE OF INFORMATION
>>>
>>> This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
>>>
>>>
>>>
>>>
>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts@febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
JA
John Ackermann N8UR
Sun, Oct 14, 2012 5:24 PM
Hi Edgardo --
I don't know a lot about the Z3805 variations other than what's on
fluke.l's site, which can be a bit confusing. All were made by (for)
HP/Agilent/Symmetricom, but there were apparently at least two
"official" variations.
Mine is the Samsung version, which has the same 10811A double oven as
the Z3801, and a 16 channel Furuno GPS, 2 x 10 MHz and 2 x 1 PPS, and
dual comm ports. It's still aging in, but it seems to perform about the
same as my older Z3801A, with stability below 1e12 from 1 second on out.
There's another telco model that has an MTI 260 oscillator, that I think
was manufactured in Korea but does not have the Samsung logo on the
front. I don't know much about that one.
Then there's the one that's boxed up like a 58503A bench unit, which I
think is not directly from HP/Agilent/Symmetricom, but is a telco unit
repackaged by someone in China. I'm not sure what's in there. One of
the ones on eBay right now says that it has a 6 channel receiver, and
was "upgraded by us which the feature is the same as 58503A". Good luck
figuring out what's in that one... (by the way, none of the current
listings are from fluke.l)
John
Edgardo Molina said the following on 10/14/2012 12:29 PM:
Dear John,
Very interesting. I haven't had the time to compare the Samsung branded units with the typical HP units. Mine is the later. It only has one communications port, plus 2x10 MHz and 2x 1PPS. As soon as I am done with the late family breakfast I will do some testing and let you know what I find.
How does the oscillators from the Samsung units compare to the HP units? Are the Samsung units superior in any sense? Both have the same GPS receiver/number of channels?
Thank you for opening my curiosity regarding this GPSDO model.
Best regards!
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 8:23 AM, John Ackermann N8UR jra@febo.com wrote:
On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
I've been using Ulrich's "z38xx" software to monitor it with good success.
John
Edgardo Molina said the following on 10/13/2012 10:18 PM:
Dear Group,
Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
- Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
- If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
- There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
Kindest regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Hi Edgardo --
I don't know a lot about the Z3805 variations other than what's on
fluke.l's site, which can be a bit confusing. All were made by (for)
HP/Agilent/Symmetricom, but there were apparently at least two
"official" variations.
Mine is the Samsung version, which has the same 10811A double oven as
the Z3801, and a 16 channel Furuno GPS, 2 x 10 MHz and 2 x 1 PPS, and
dual comm ports. It's still aging in, but it seems to perform about the
same as my older Z3801A, with stability below 1e12 from 1 second on out.
There's another telco model that has an MTI 260 oscillator, that I think
was manufactured in Korea but does not have the Samsung logo on the
front. I don't know much about that one.
Then there's the one that's boxed up like a 58503A bench unit, which I
think is not directly from HP/Agilent/Symmetricom, but is a telco unit
repackaged by someone in China. I'm not sure what's in there. One of
the ones on eBay right now says that it has a 6 channel receiver, and
was "upgraded by us which the feature is the same as 58503A". Good luck
figuring out what's in that one... (by the way, none of the current
listings are from fluke.l)
John
----
Edgardo Molina said the following on 10/14/2012 12:29 PM:
> Dear John,
>
> Very interesting. I haven't had the time to compare the Samsung branded units with the typical HP units. Mine is the later. It only has one communications port, plus 2x10 MHz and 2x 1PPS. As soon as I am done with the late family breakfast I will do some testing and let you know what I find.
>
> How does the oscillators from the Samsung units compare to the HP units? Are the Samsung units superior in any sense? Both have the same GPS receiver/number of channels?
>
> Thank you for opening my curiosity regarding this GPSDO model.
>
> Best regards!
>
>
>
> Edgardo Molina
> Dirección IPTEL
>
> www.iptel.net.mx
>
> T : 55 55 55202444
> M : 04455 20501854
>
> Piensa en Bits SA de CV
>
>
>
> Información anexa:
>
>
>
>
> CONFIDENCIALIDAD DE INFORMACION
>
> Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
>
>
> NON-DISCLOSURE OF INFORMATION
>
> This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
>
>
>
>
>
> On Oct 14, 2012, at 8:23 AM, John Ackermann N8UR <jra@febo.com> wrote:
>
>> On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
>>
>> I've been using Ulrich's "z38xx" software to monitor it with good success.
>>
>> John
>> ----
>>
>> Edgardo Molina said the following on 10/13/2012 10:18 PM:
>>> Dear Group,
>>>
>>> Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
>>>
>>> After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
>>>
>>> 1. Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
>>> 2. If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
>>> 3. There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
>>>
>>> Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
>>>
>>> Kindest regards,
>>>
>>>
>>>
>>> Edgardo Molina
>>> Dirección IPTEL
>>>
>>> www.iptel.net.mx
>>>
>>> T : 55 55 55202444
>>> M : 04455 20501854
>>>
>>> Piensa en Bits SA de CV
>>>
>>>
>>>
>>> Información anexa:
>>>
>>>
>>>
>>>
>>> CONFIDENCIALIDAD DE INFORMACION
>>>
>>> Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
>>>
>>>
>>> NON-DISCLOSURE OF INFORMATION
>>>
>>> This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
>>>
>>>
>>>
>>>
>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts@febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
EM
Edgardo Molina
Sun, Oct 14, 2012 5:45 PM
Dear John,
Thank you for your explanation on the Z3805A models. Mine is identical to the 58305A. Same case, same form factor, same rubber guard ends. HP logos and design all around. I haven't opened them yet. They still have calibration seals on top of the screw head case. As I find time to do it. I will let you know about the inner workings.
I will take a look at the current online offerings to also see if they match my units. Thanks again.
Regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 12:24 PM, John Ackermann N8UR jra@febo.com wrote:
Hi Edgardo --
I don't know a lot about the Z3805 variations other than what's on fluke.l's site, which can be a bit confusing. All were made by (for) HP/Agilent/Symmetricom, but there were apparently at least two "official" variations.
Mine is the Samsung version, which has the same 10811A double oven as the Z3801, and a 16 channel Furuno GPS, 2 x 10 MHz and 2 x 1 PPS, and dual comm ports. It's still aging in, but it seems to perform about the same as my older Z3801A, with stability below 1e12 from 1 second on out.
There's another telco model that has an MTI 260 oscillator, that I think was manufactured in Korea but does not have the Samsung logo on the front. I don't know much about that one.
Then there's the one that's boxed up like a 58503A bench unit, which I think is not directly from HP/Agilent/Symmetricom, but is a telco unit repackaged by someone in China. I'm not sure what's in there. One of the ones on eBay right now says that it has a 6 channel receiver, and was "upgraded by us which the feature is the same as 58503A". Good luck figuring out what's in that one... (by the way, none of the current listings are from fluke.l)
John
Edgardo Molina said the following on 10/14/2012 12:29 PM:
Dear John,
Very interesting. I haven't had the time to compare the Samsung branded units with the typical HP units. Mine is the later. It only has one communications port, plus 2x10 MHz and 2x 1PPS. As soon as I am done with the late family breakfast I will do some testing and let you know what I find.
How does the oscillators from the Samsung units compare to the HP units? Are the Samsung units superior in any sense? Both have the same GPS receiver/number of channels?
Thank you for opening my curiosity regarding this GPSDO model.
Best regards!
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 8:23 AM, John Ackermann N8UR jra@febo.com wrote:
On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
I've been using Ulrich's "z38xx" software to monitor it with good success.
John
Edgardo Molina said the following on 10/13/2012 10:18 PM:
Dear Group,
Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
- Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
- If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
- There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
Kindest regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Dear John,
Thank you for your explanation on the Z3805A models. Mine is identical to the 58305A. Same case, same form factor, same rubber guard ends. HP logos and design all around. I haven't opened them yet. They still have calibration seals on top of the screw head case. As I find time to do it. I will let you know about the inner workings.
I will take a look at the current online offerings to also see if they match my units. Thanks again.
Regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 12:24 PM, John Ackermann N8UR <jra@febo.com> wrote:
> Hi Edgardo --
>
> I don't know a lot about the Z3805 variations other than what's on fluke.l's site, which can be a bit confusing. All were made by (for) HP/Agilent/Symmetricom, but there were apparently at least two "official" variations.
>
> Mine is the Samsung version, which has the same 10811A double oven as the Z3801, and a 16 channel Furuno GPS, 2 x 10 MHz and 2 x 1 PPS, and dual comm ports. It's still aging in, but it seems to perform about the same as my older Z3801A, with stability below 1e12 from 1 second on out.
>
> There's another telco model that has an MTI 260 oscillator, that I think was manufactured in Korea but does not have the Samsung logo on the front. I don't know much about that one.
>
> Then there's the one that's boxed up like a 58503A bench unit, which I think is not directly from HP/Agilent/Symmetricom, but is a telco unit repackaged by someone in China. I'm not sure what's in there. One of the ones on eBay right now says that it has a 6 channel receiver, and was "upgraded by us which the feature is the same as 58503A". Good luck figuring out what's in that one... (by the way, none of the current listings are from fluke.l)
>
> John
> ----
>
> Edgardo Molina said the following on 10/14/2012 12:29 PM:
>> Dear John,
>>
>> Very interesting. I haven't had the time to compare the Samsung branded units with the typical HP units. Mine is the later. It only has one communications port, plus 2x10 MHz and 2x 1PPS. As soon as I am done with the late family breakfast I will do some testing and let you know what I find.
>>
>> How does the oscillators from the Samsung units compare to the HP units? Are the Samsung units superior in any sense? Both have the same GPS receiver/number of channels?
>>
>> Thank you for opening my curiosity regarding this GPSDO model.
>>
>> Best regards!
>>
>>
>>
>> Edgardo Molina
>> Dirección IPTEL
>>
>> www.iptel.net.mx
>>
>> T : 55 55 55202444
>> M : 04455 20501854
>>
>> Piensa en Bits SA de CV
>>
>>
>>
>> Información anexa:
>>
>>
>>
>>
>> CONFIDENCIALIDAD DE INFORMACION
>>
>> Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
>>
>>
>> NON-DISCLOSURE OF INFORMATION
>>
>> This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
>>
>>
>>
>>
>>
>> On Oct 14, 2012, at 8:23 AM, John Ackermann N8UR <jra@febo.com> wrote:
>>
>>> On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
>>>
>>> I've been using Ulrich's "z38xx" software to monitor it with good success.
>>>
>>> John
>>> ----
>>>
>>> Edgardo Molina said the following on 10/13/2012 10:18 PM:
>>>> Dear Group,
>>>>
>>>> Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
>>>>
>>>> After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
>>>>
>>>> 1. Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
>>>> 2. If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
>>>> 3. There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
>>>>
>>>> Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
>>>>
>>>> Kindest regards,
>>>>
>>>>
>>>>
>>>> Edgardo Molina
>>>> Dirección IPTEL
>>>>
>>>> www.iptel.net.mx
>>>>
>>>> T : 55 55 55202444
>>>> M : 04455 20501854
>>>>
>>>> Piensa en Bits SA de CV
>>>>
>>>>
>>>>
>>>> Información anexa:
>>>>
>>>>
>>>>
>>>>
>>>> CONFIDENCIALIDAD DE INFORMACION
>>>>
>>>> Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
>>>>
>>>>
>>>> NON-DISCLOSURE OF INFORMATION
>>>>
>>>> This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
>>>>
>>>>
>>>>
>>>>
>>>>
>>>> _______________________________________________
>>>> time-nuts mailing list -- time-nuts@febo.com
>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>> and follow the instructions there.
>>>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
EM
Edgardo Molina
Sun, Oct 14, 2012 6:18 PM
Dear John,
I just opened one box. Here is what I found inside:
HP-10811-60165 double oven oscillator
Furuno 16 Ch GPS receiver
Now I will check if communications port is either RS-232 or RS-422 as per Azelio advise. Once done, I will try Chuck's monitoring and command software.
My wife and kids are getting impatient. Ringing my lab's doorbell. Lunch time! We're going some sushi place.
The afternoon is mine to continue testing and finishing my thesis translation. I am fortunate that tomorrow Dr. Levine is giving some advise on it at CENAM.
Until then…
Cheers!
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 12:24 PM, John Ackermann N8UR jra@febo.com wrote:
Hi Edgardo --
I don't know a lot about the Z3805 variations other than what's on fluke.l's site, which can be a bit confusing. All were made by (for) HP/Agilent/Symmetricom, but there were apparently at least two "official" variations.
Mine is the Samsung version, which has the same 10811A double oven as the Z3801, and a 16 channel Furuno GPS, 2 x 10 MHz and 2 x 1 PPS, and dual comm ports. It's still aging in, but it seems to perform about the same as my older Z3801A, with stability below 1e12 from 1 second on out.
There's another telco model that has an MTI 260 oscillator, that I think was manufactured in Korea but does not have the Samsung logo on the front. I don't know much about that one.
Then there's the one that's boxed up like a 58503A bench unit, which I think is not directly from HP/Agilent/Symmetricom, but is a telco unit repackaged by someone in China. I'm not sure what's in there. One of the ones on eBay right now says that it has a 6 channel receiver, and was "upgraded by us which the feature is the same as 58503A". Good luck figuring out what's in that one... (by the way, none of the current listings are from fluke.l)
John
Edgardo Molina said the following on 10/14/2012 12:29 PM:
Dear John,
Very interesting. I haven't had the time to compare the Samsung branded units with the typical HP units. Mine is the later. It only has one communications port, plus 2x10 MHz and 2x 1PPS. As soon as I am done with the late family breakfast I will do some testing and let you know what I find.
How does the oscillators from the Samsung units compare to the HP units? Are the Samsung units superior in any sense? Both have the same GPS receiver/number of channels?
Thank you for opening my curiosity regarding this GPSDO model.
Best regards!
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 8:23 AM, John Ackermann N8UR jra@febo.com wrote:
On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
I've been using Ulrich's "z38xx" software to monitor it with good success.
John
Edgardo Molina said the following on 10/13/2012 10:18 PM:
Dear Group,
Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
- Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
- If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
- There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
Kindest regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Dear John,
I just opened one box. Here is what I found inside:
HP-10811-60165 double oven oscillator
Furuno 16 Ch GPS receiver
Now I will check if communications port is either RS-232 or RS-422 as per Azelio advise. Once done, I will try Chuck's monitoring and command software.
My wife and kids are getting impatient. Ringing my lab's doorbell. Lunch time! We're going some sushi place.
The afternoon is mine to continue testing and finishing my thesis translation. I am fortunate that tomorrow Dr. Levine is giving some advise on it at CENAM.
Until then…
Cheers!
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 12:24 PM, John Ackermann N8UR <jra@febo.com> wrote:
> Hi Edgardo --
>
> I don't know a lot about the Z3805 variations other than what's on fluke.l's site, which can be a bit confusing. All were made by (for) HP/Agilent/Symmetricom, but there were apparently at least two "official" variations.
>
> Mine is the Samsung version, which has the same 10811A double oven as the Z3801, and a 16 channel Furuno GPS, 2 x 10 MHz and 2 x 1 PPS, and dual comm ports. It's still aging in, but it seems to perform about the same as my older Z3801A, with stability below 1e12 from 1 second on out.
>
> There's another telco model that has an MTI 260 oscillator, that I think was manufactured in Korea but does not have the Samsung logo on the front. I don't know much about that one.
>
> Then there's the one that's boxed up like a 58503A bench unit, which I think is not directly from HP/Agilent/Symmetricom, but is a telco unit repackaged by someone in China. I'm not sure what's in there. One of the ones on eBay right now says that it has a 6 channel receiver, and was "upgraded by us which the feature is the same as 58503A". Good luck figuring out what's in that one... (by the way, none of the current listings are from fluke.l)
>
> John
> ----
>
> Edgardo Molina said the following on 10/14/2012 12:29 PM:
>> Dear John,
>>
>> Very interesting. I haven't had the time to compare the Samsung branded units with the typical HP units. Mine is the later. It only has one communications port, plus 2x10 MHz and 2x 1PPS. As soon as I am done with the late family breakfast I will do some testing and let you know what I find.
>>
>> How does the oscillators from the Samsung units compare to the HP units? Are the Samsung units superior in any sense? Both have the same GPS receiver/number of channels?
>>
>> Thank you for opening my curiosity regarding this GPSDO model.
>>
>> Best regards!
>>
>>
>>
>> Edgardo Molina
>> Dirección IPTEL
>>
>> www.iptel.net.mx
>>
>> T : 55 55 55202444
>> M : 04455 20501854
>>
>> Piensa en Bits SA de CV
>>
>>
>>
>> Información anexa:
>>
>>
>>
>>
>> CONFIDENCIALIDAD DE INFORMACION
>>
>> Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
>>
>>
>> NON-DISCLOSURE OF INFORMATION
>>
>> This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
>>
>>
>>
>>
>>
>> On Oct 14, 2012, at 8:23 AM, John Ackermann N8UR <jra@febo.com> wrote:
>>
>>> On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
>>>
>>> I've been using Ulrich's "z38xx" software to monitor it with good success.
>>>
>>> John
>>> ----
>>>
>>> Edgardo Molina said the following on 10/13/2012 10:18 PM:
>>>> Dear Group,
>>>>
>>>> Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
>>>>
>>>> After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
>>>>
>>>> 1. Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
>>>> 2. If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
>>>> 3. There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
>>>>
>>>> Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
>>>>
>>>> Kindest regards,
>>>>
>>>>
>>>>
>>>> Edgardo Molina
>>>> Dirección IPTEL
>>>>
>>>> www.iptel.net.mx
>>>>
>>>> T : 55 55 55202444
>>>> M : 04455 20501854
>>>>
>>>> Piensa en Bits SA de CV
>>>>
>>>>
>>>>
>>>> Información anexa:
>>>>
>>>>
>>>>
>>>>
>>>> CONFIDENCIALIDAD DE INFORMACION
>>>>
>>>> Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
>>>>
>>>>
>>>> NON-DISCLOSURE OF INFORMATION
>>>>
>>>> This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
>>>>
>>>>
>>>>
>>>>
>>>>
>>>> _______________________________________________
>>>> time-nuts mailing list -- time-nuts@febo.com
>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>> and follow the instructions there.
>>>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
SJ
Said Jackson
Sun, Oct 14, 2012 10:15 PM
John,
I believe Edgardo bought the reprogrammed 58503A unit (was a Z3805A before) from Ebay.
I have the same units recently purchased from yinxhu(sp?) on Ebay and Edgardo just confirmed these are indeed 16 channel units with true 10811 Docxos.
Explains why they work so great..
Edgardo, I use GPSCon with these, and mine are all Rs232..
Mine achieve performance in the xE-013 and noise below -162dBc (Thunderbolt eat your heart out :). I believe there is no better GPSDO out there presently, much less available for sale and at those prices. PN at 1Hz is better than -105dBc..
I regularly get below 1us predicted drift as shown by the syst:stat? command.
Bye,
Said
On Oct 14, 2012, at 10:24, John Ackermann N8UR jra@febo.com wrote:
Hi Edgardo --
I don't know a lot about the Z3805 variations other than what's on fluke.l's site, which can be a bit confusing. All were made by (for) HP/Agilent/Symmetricom, but there were apparently at least two "official" variations.
Mine is the Samsung version, which has the same 10811A double oven as the Z3801, and a 16 channel Furuno GPS, 2 x 10 MHz and 2 x 1 PPS, and dual comm ports. It's still aging in, but it seems to perform about the same as my older Z3801A, with stability below 1e12 from 1 second on out.
There's another telco model that has an MTI 260 oscillator, that I think was manufactured in Korea but does not have the Samsung logo on the front. I don't know much about that one.
Then there's the one that's boxed up like a 58503A bench unit, which I think is not directly from HP/Agilent/Symmetricom, but is a telco unit repackaged by someone in China. I'm not sure what's in there. One of the ones on eBay right now says that it has a 6 channel receiver, and was "upgraded by us which the feature is the same as 58503A". Good luck figuring out what's in that one... (by the way, none of the current listings are from fluke.l)
John
Edgardo Molina said the following on 10/14/2012 12:29 PM:
Dear John,
Very interesting. I haven't had the time to compare the Samsung branded units with the typical HP units. Mine is the later. It only has one communications port, plus 2x10 MHz and 2x 1PPS. As soon as I am done with the late family breakfast I will do some testing and let you know what I find.
How does the oscillators from the Samsung units compare to the HP units? Are the Samsung units superior in any sense? Both have the same GPS receiver/number of channels?
Thank you for opening my curiosity regarding this GPSDO model.
Best regards!
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 8:23 AM, John Ackermann N8UR jra@febo.com wrote:
On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
I've been using Ulrich's "z38xx" software to monitor it with good success.
John
Edgardo Molina said the following on 10/13/2012 10:18 PM:
Dear Group,
Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
- Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
- If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
- There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
Kindest regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
John,
I believe Edgardo bought the reprogrammed 58503A unit (was a Z3805A before) from Ebay.
I have the same units recently purchased from yinxhu(sp?) on Ebay and Edgardo just confirmed these are indeed 16 channel units with true 10811 Docxos.
Explains why they work so great..
Edgardo, I use GPSCon with these, and mine are all Rs232..
Mine achieve performance in the xE-013 and noise below -162dBc (Thunderbolt eat your heart out :). I believe there is no better GPSDO out there presently, much less available for sale and at those prices. PN at 1Hz is better than -105dBc..
I regularly get below 1us predicted drift as shown by the syst:stat? command.
Bye,
Said
On Oct 14, 2012, at 10:24, John Ackermann N8UR <jra@febo.com> wrote:
> Hi Edgardo --
>
> I don't know a lot about the Z3805 variations other than what's on fluke.l's site, which can be a bit confusing. All were made by (for) HP/Agilent/Symmetricom, but there were apparently at least two "official" variations.
>
> Mine is the Samsung version, which has the same 10811A double oven as the Z3801, and a 16 channel Furuno GPS, 2 x 10 MHz and 2 x 1 PPS, and dual comm ports. It's still aging in, but it seems to perform about the same as my older Z3801A, with stability below 1e12 from 1 second on out.
>
> There's another telco model that has an MTI 260 oscillator, that I think was manufactured in Korea but does not have the Samsung logo on the front. I don't know much about that one.
>
> Then there's the one that's boxed up like a 58503A bench unit, which I think is not directly from HP/Agilent/Symmetricom, but is a telco unit repackaged by someone in China. I'm not sure what's in there. One of the ones on eBay right now says that it has a 6 channel receiver, and was "upgraded by us which the feature is the same as 58503A". Good luck figuring out what's in that one... (by the way, none of the current listings are from fluke.l)
>
> John
> ----
>
> Edgardo Molina said the following on 10/14/2012 12:29 PM:
>> Dear John,
>>
>> Very interesting. I haven't had the time to compare the Samsung branded units with the typical HP units. Mine is the later. It only has one communications port, plus 2x10 MHz and 2x 1PPS. As soon as I am done with the late family breakfast I will do some testing and let you know what I find.
>>
>> How does the oscillators from the Samsung units compare to the HP units? Are the Samsung units superior in any sense? Both have the same GPS receiver/number of channels?
>>
>> Thank you for opening my curiosity regarding this GPSDO model.
>>
>> Best regards!
>>
>>
>>
>> Edgardo Molina
>> Dirección IPTEL
>>
>> www.iptel.net.mx
>>
>> T : 55 55 55202444
>> M : 04455 20501854
>>
>> Piensa en Bits SA de CV
>>
>>
>>
>> Información anexa:
>>
>>
>>
>>
>> CONFIDENCIALIDAD DE INFORMACION
>>
>> Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
>>
>>
>> NON-DISCLOSURE OF INFORMATION
>>
>> This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
>>
>>
>>
>>
>>
>> On Oct 14, 2012, at 8:23 AM, John Ackermann N8UR <jra@febo.com> wrote:
>>
>>> On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
>>>
>>> I've been using Ulrich's "z38xx" software to monitor it with good success.
>>>
>>> John
>>> ----
>>>
>>> Edgardo Molina said the following on 10/13/2012 10:18 PM:
>>>> Dear Group,
>>>>
>>>> Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
>>>>
>>>> After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
>>>>
>>>> 1. Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
>>>> 2. If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
>>>> 3. There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
>>>>
>>>> Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
>>>>
>>>> Kindest regards,
>>>>
>>>>
>>>>
>>>> Edgardo Molina
>>>> Dirección IPTEL
>>>>
>>>> www.iptel.net.mx
>>>>
>>>> T : 55 55 55202444
>>>> M : 04455 20501854
>>>>
>>>> Piensa en Bits SA de CV
>>>>
>>>>
>>>>
>>>> Información anexa:
>>>>
>>>>
>>>>
>>>>
>>>> CONFIDENCIALIDAD DE INFORMACION
>>>>
>>>> Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
>>>>
>>>>
>>>> NON-DISCLOSURE OF INFORMATION
>>>>
>>>> This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
>>>>
>>>>
>>>>
>>>>
>>>>
>>>> _______________________________________________
>>>> time-nuts mailing list -- time-nuts@febo.com
>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>> and follow the instructions there.
>>>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
EM
Edgardo Molina
Sun, Oct 14, 2012 10:40 PM
Dear Said,
Thank you. It seems I ended up buying the same units and from the same provider. I recall you mentioning before I got mine about the availability of those models in the Asia-Pacific region. That explains clearly the capabilities and quality of the units. I just ran the tests Azelio suggested and in fact it is an RS-232 port. Thank you also for confirming the ports are RS-232 from your side. Now I am connecting one of the receivers to the PC and start playing around with it.
Derived question: Will I get better GPSDO performance if I use a 58503A receiver? Or do you think our units match those 58503A GPSDOs out there?
Again, thanks to everybody who helped me to figure out about my receivers and their inner workings. As always, you all made my day.
Best regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 5:15 PM, Said Jackson saidjack@aol.com wrote:
John,
I believe Edgardo bought the reprogrammed 58503A unit (was a Z3805A before) from Ebay.
I have the same units recently purchased from yinxhu(sp?) on Ebay and Edgardo just confirmed these are indeed 16 channel units with true 10811 Docxos.
Explains why they work so great..
Edgardo, I use GPSCon with these, and mine are all Rs232..
Mine achieve performance in the xE-013 and noise below -162dBc (Thunderbolt eat your heart out :). I believe there is no better GPSDO out there presently, much less available for sale and at those prices. PN at 1Hz is better than -105dBc..
I regularly get below 1us predicted drift as shown by the syst:stat? command.
Bye,
Said
On Oct 14, 2012, at 10:24, John Ackermann N8UR jra@febo.com wrote:
Hi Edgardo --
I don't know a lot about the Z3805 variations other than what's on fluke.l's site, which can be a bit confusing. All were made by (for) HP/Agilent/Symmetricom, but there were apparently at least two "official" variations.
Mine is the Samsung version, which has the same 10811A double oven as the Z3801, and a 16 channel Furuno GPS, 2 x 10 MHz and 2 x 1 PPS, and dual comm ports. It's still aging in, but it seems to perform about the same as my older Z3801A, with stability below 1e12 from 1 second on out.
There's another telco model that has an MTI 260 oscillator, that I think was manufactured in Korea but does not have the Samsung logo on the front. I don't know much about that one.
Then there's the one that's boxed up like a 58503A bench unit, which I think is not directly from HP/Agilent/Symmetricom, but is a telco unit repackaged by someone in China. I'm not sure what's in there. One of the ones on eBay right now says that it has a 6 channel receiver, and was "upgraded by us which the feature is the same as 58503A". Good luck figuring out what's in that one... (by the way, none of the current listings are from fluke.l)
John
Edgardo Molina said the following on 10/14/2012 12:29 PM:
Dear John,
Very interesting. I haven't had the time to compare the Samsung branded units with the typical HP units. Mine is the later. It only has one communications port, plus 2x10 MHz and 2x 1PPS. As soon as I am done with the late family breakfast I will do some testing and let you know what I find.
How does the oscillators from the Samsung units compare to the HP units? Are the Samsung units superior in any sense? Both have the same GPS receiver/number of channels?
Thank you for opening my curiosity regarding this GPSDO model.
Best regards!
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 8:23 AM, John Ackermann N8UR jra@febo.com wrote:
On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
I've been using Ulrich's "z38xx" software to monitor it with good success.
John
Edgardo Molina said the following on 10/13/2012 10:18 PM:
Dear Group,
Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
- Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
- If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
- There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
Kindest regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Dear Said,
Thank you. It seems I ended up buying the same units and from the same provider. I recall you mentioning before I got mine about the availability of those models in the Asia-Pacific region. That explains clearly the capabilities and quality of the units. I just ran the tests Azelio suggested and in fact it is an RS-232 port. Thank you also for confirming the ports are RS-232 from your side. Now I am connecting one of the receivers to the PC and start playing around with it.
Derived question: Will I get better GPSDO performance if I use a 58503A receiver? Or do you think our units match those 58503A GPSDOs out there?
Again, thanks to everybody who helped me to figure out about my receivers and their inner workings. As always, you all made my day.
Best regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 5:15 PM, Said Jackson <saidjack@aol.com> wrote:
> John,
>
> I believe Edgardo bought the reprogrammed 58503A unit (was a Z3805A before) from Ebay.
>
> I have the same units recently purchased from yinxhu(sp?) on Ebay and Edgardo just confirmed these are indeed 16 channel units with true 10811 Docxos.
>
> Explains why they work so great..
>
> Edgardo, I use GPSCon with these, and mine are all Rs232..
>
> Mine achieve performance in the xE-013 and noise below -162dBc (Thunderbolt eat your heart out :). I believe there is no better GPSDO out there presently, much less available for sale and at those prices. PN at 1Hz is better than -105dBc..
>
> I regularly get below 1us predicted drift as shown by the syst:stat? command.
>
> Bye,
> Said
>
>
>
> On Oct 14, 2012, at 10:24, John Ackermann N8UR <jra@febo.com> wrote:
>
>> Hi Edgardo --
>>
>> I don't know a lot about the Z3805 variations other than what's on fluke.l's site, which can be a bit confusing. All were made by (for) HP/Agilent/Symmetricom, but there were apparently at least two "official" variations.
>>
>> Mine is the Samsung version, which has the same 10811A double oven as the Z3801, and a 16 channel Furuno GPS, 2 x 10 MHz and 2 x 1 PPS, and dual comm ports. It's still aging in, but it seems to perform about the same as my older Z3801A, with stability below 1e12 from 1 second on out.
>>
>> There's another telco model that has an MTI 260 oscillator, that I think was manufactured in Korea but does not have the Samsung logo on the front. I don't know much about that one.
>>
>> Then there's the one that's boxed up like a 58503A bench unit, which I think is not directly from HP/Agilent/Symmetricom, but is a telco unit repackaged by someone in China. I'm not sure what's in there. One of the ones on eBay right now says that it has a 6 channel receiver, and was "upgraded by us which the feature is the same as 58503A". Good luck figuring out what's in that one... (by the way, none of the current listings are from fluke.l)
>>
>> John
>> ----
>>
>> Edgardo Molina said the following on 10/14/2012 12:29 PM:
>>> Dear John,
>>>
>>> Very interesting. I haven't had the time to compare the Samsung branded units with the typical HP units. Mine is the later. It only has one communications port, plus 2x10 MHz and 2x 1PPS. As soon as I am done with the late family breakfast I will do some testing and let you know what I find.
>>>
>>> How does the oscillators from the Samsung units compare to the HP units? Are the Samsung units superior in any sense? Both have the same GPS receiver/number of channels?
>>>
>>> Thank you for opening my curiosity regarding this GPSDO model.
>>>
>>> Best regards!
>>>
>>>
>>>
>>> Edgardo Molina
>>> Dirección IPTEL
>>>
>>> www.iptel.net.mx
>>>
>>> T : 55 55 55202444
>>> M : 04455 20501854
>>>
>>> Piensa en Bits SA de CV
>>>
>>>
>>>
>>> Información anexa:
>>>
>>>
>>>
>>>
>>> CONFIDENCIALIDAD DE INFORMACION
>>>
>>> Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
>>>
>>>
>>> NON-DISCLOSURE OF INFORMATION
>>>
>>> This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
>>>
>>>
>>>
>>>
>>>
>>> On Oct 14, 2012, at 8:23 AM, John Ackermann N8UR <jra@febo.com> wrote:
>>>
>>>> On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
>>>>
>>>> I've been using Ulrich's "z38xx" software to monitor it with good success.
>>>>
>>>> John
>>>> ----
>>>>
>>>> Edgardo Molina said the following on 10/13/2012 10:18 PM:
>>>>> Dear Group,
>>>>>
>>>>> Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
>>>>>
>>>>> After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
>>>>>
>>>>> 1. Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
>>>>> 2. If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
>>>>> 3. There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
>>>>>
>>>>> Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
>>>>>
>>>>> Kindest regards,
>>>>>
>>>>>
>>>>>
>>>>> Edgardo Molina
>>>>> Dirección IPTEL
>>>>>
>>>>> www.iptel.net.mx
>>>>>
>>>>> T : 55 55 55202444
>>>>> M : 04455 20501854
>>>>>
>>>>> Piensa en Bits SA de CV
>>>>>
>>>>>
>>>>>
>>>>> Información anexa:
>>>>>
>>>>>
>>>>>
>>>>>
>>>>> CONFIDENCIALIDAD DE INFORMACION
>>>>>
>>>>> Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
>>>>>
>>>>>
>>>>> NON-DISCLOSURE OF INFORMATION
>>>>>
>>>>> This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
>>>>>
>>>>>
>>>>>
>>>>>
>>>>>
>>>>> _______________________________________________
>>>>> time-nuts mailing list -- time-nuts@febo.com
>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>> and follow the instructions there.
>>>>>
>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts@febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
JG
Joseph Gray
Sun, Oct 14, 2012 11:19 PM
Just cuious, of the two 58503A versions that yixunhk sells (Z3801A and
Z3805A), which one has the better performance? I see that the Z3805A
version has two 10 MHz outputs and one PPS.
Joe Gray
W5JG
On Sun, Oct 14, 2012 at 4:15 PM, Said Jackson saidjack@aol.com wrote:
John,
I believe Edgardo bought the reprogrammed 58503A unit (was a Z3805A before) from Ebay.
I have the same units recently purchased from yinxhu(sp?) on Ebay and Edgardo just confirmed these are indeed 16 channel units with true 10811 Docxos.
Explains why they work so great..
Edgardo, I use GPSCon with these, and mine are all Rs232..
Mine achieve performance in the xE-013 and noise below -162dBc (Thunderbolt eat your heart out :). I believe there is no better GPSDO out there presently, much less available for sale and at those prices. PN at 1Hz is better than -105dBc..
I regularly get below 1us predicted drift as shown by the syst:stat? command.
Bye,
Said
On Oct 14, 2012, at 10:24, John Ackermann N8UR jra@febo.com wrote:
Hi Edgardo --
I don't know a lot about the Z3805 variations other than what's on fluke.l's site, which can be a bit confusing. All were made by (for) HP/Agilent/Symmetricom, but there were apparently at least two "official" variations.
Mine is the Samsung version, which has the same 10811A double oven as the Z3801, and a 16 channel Furuno GPS, 2 x 10 MHz and 2 x 1 PPS, and dual comm ports. It's still aging in, but it seems to perform about the same as my older Z3801A, with stability below 1e12 from 1 second on out.
There's another telco model that has an MTI 260 oscillator, that I think was manufactured in Korea but does not have the Samsung logo on the front. I don't know much about that one.
Then there's the one that's boxed up like a 58503A bench unit, which I think is not directly from HP/Agilent/Symmetricom, but is a telco unit repackaged by someone in China. I'm not sure what's in there. One of the ones on eBay right now says that it has a 6 channel receiver, and was "upgraded by us which the feature is the same as 58503A". Good luck figuring out what's in that one... (by the way, none of the current listings are from fluke.l)
John
Edgardo Molina said the following on 10/14/2012 12:29 PM:
Dear John,
Very interesting. I haven't had the time to compare the Samsung branded units with the typical HP units. Mine is the later. It only has one communications port, plus 2x10 MHz and 2x 1PPS. As soon as I am done with the late family breakfast I will do some testing and let you know what I find.
How does the oscillators from the Samsung units compare to the HP units? Are the Samsung units superior in any sense? Both have the same GPS receiver/number of channels?
Thank you for opening my curiosity regarding this GPSDO model.
Best regards!
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 14, 2012, at 8:23 AM, John Ackermann N8UR jra@febo.com wrote:
On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
I've been using Ulrich's "z38xx" software to monitor it with good success.
John
Edgardo Molina said the following on 10/13/2012 10:18 PM:
Dear Group,
Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
- Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
- If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
- There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
Kindest regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Just cuious, of the two 58503A versions that yixunhk sells (Z3801A and
Z3805A), which one has the better performance? I see that the Z3805A
version has two 10 MHz outputs and one PPS.
Joe Gray
W5JG
On Sun, Oct 14, 2012 at 4:15 PM, Said Jackson <saidjack@aol.com> wrote:
> John,
>
> I believe Edgardo bought the reprogrammed 58503A unit (was a Z3805A before) from Ebay.
>
> I have the same units recently purchased from yinxhu(sp?) on Ebay and Edgardo just confirmed these are indeed 16 channel units with true 10811 Docxos.
>
> Explains why they work so great..
>
> Edgardo, I use GPSCon with these, and mine are all Rs232..
>
> Mine achieve performance in the xE-013 and noise below -162dBc (Thunderbolt eat your heart out :). I believe there is no better GPSDO out there presently, much less available for sale and at those prices. PN at 1Hz is better than -105dBc..
>
> I regularly get below 1us predicted drift as shown by the syst:stat? command.
>
> Bye,
> Said
>
>
>
> On Oct 14, 2012, at 10:24, John Ackermann N8UR <jra@febo.com> wrote:
>
>> Hi Edgardo --
>>
>> I don't know a lot about the Z3805 variations other than what's on fluke.l's site, which can be a bit confusing. All were made by (for) HP/Agilent/Symmetricom, but there were apparently at least two "official" variations.
>>
>> Mine is the Samsung version, which has the same 10811A double oven as the Z3801, and a 16 channel Furuno GPS, 2 x 10 MHz and 2 x 1 PPS, and dual comm ports. It's still aging in, but it seems to perform about the same as my older Z3801A, with stability below 1e12 from 1 second on out.
>>
>> There's another telco model that has an MTI 260 oscillator, that I think was manufactured in Korea but does not have the Samsung logo on the front. I don't know much about that one.
>>
>> Then there's the one that's boxed up like a 58503A bench unit, which I think is not directly from HP/Agilent/Symmetricom, but is a telco unit repackaged by someone in China. I'm not sure what's in there. One of the ones on eBay right now says that it has a 6 channel receiver, and was "upgraded by us which the feature is the same as 58503A". Good luck figuring out what's in that one... (by the way, none of the current listings are from fluke.l)
>>
>> John
>> ----
>>
>> Edgardo Molina said the following on 10/14/2012 12:29 PM:
>>> Dear John,
>>>
>>> Very interesting. I haven't had the time to compare the Samsung branded units with the typical HP units. Mine is the later. It only has one communications port, plus 2x10 MHz and 2x 1PPS. As soon as I am done with the late family breakfast I will do some testing and let you know what I find.
>>>
>>> How does the oscillators from the Samsung units compare to the HP units? Are the Samsung units superior in any sense? Both have the same GPS receiver/number of channels?
>>>
>>> Thank you for opening my curiosity regarding this GPSDO model.
>>>
>>> Best regards!
>>>
>>>
>>>
>>> Edgardo Molina
>>> Dirección IPTEL
>>>
>>> www.iptel.net.mx
>>>
>>> T : 55 55 55202444
>>> M : 04455 20501854
>>>
>>> Piensa en Bits SA de CV
>>>
>>>
>>>
>>> Información anexa:
>>>
>>>
>>>
>>>
>>> CONFIDENCIALIDAD DE INFORMACION
>>>
>>> Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
>>>
>>>
>>> NON-DISCLOSURE OF INFORMATION
>>>
>>> This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
>>>
>>>
>>>
>>>
>>>
>>> On Oct 14, 2012, at 8:23 AM, John Ackermann N8UR <jra@febo.com> wrote:
>>>
>>>> On the unit I bought from fluke.l, with a "Samsung" label, there are two serial ports. The bottom one is the control port and is definitely RS-232. The top port has only three wires (TXD, RXD, SG) and I think is used for some sort of monitoring.
>>>>
>>>> I've been using Ulrich's "z38xx" software to monitor it with good success.
>>>>
>>>> John
>>>> ----
>>>>
>>>> Edgardo Molina said the following on 10/13/2012 10:18 PM:
>>>>> Dear Group,
>>>>>
>>>>> Good evening. I bought two HP Z3805A units a few weeks ago. They were bare units without antenna or power supply. After struggling to find a suitable 24V 3A noise free power supply for each of them, I mounted a couple of Symmetricom antennas and just fired them up.
>>>>>
>>>>> After around 25 minutes, the GPS Lock light lighted up. As the manual for this particular model is unobtainable in several searches through the Internet, I found that the readily available manual for the Z3801A would cover much of the Z3905A operation. I also got a software for monitoring the operation from BD Systems Inc. In Colorado. Now that I am sure both are operating and locked only by looking at the front panel lights, I would like to ask a couple of questions regarding their communication to the PC.
>>>>>
>>>>> 1. Is the DB-25 serial port carrying RS-232 signals? I see the port of the Z380A1 is different: RS-422.
>>>>> 2. If the port is plain RS-232 is it possible to use a Serial to USB adapter? I have heard about the inconsistencies that such an adapter could cause when used with a TBolt. Should I assume this is the same case with the Z3805A?
>>>>> 3. There was once a project for building a TBolt state monitor on a 2 line LCD screen. I own three of those monitors and they are just great. Is there an equivalent for these HP GPS units?
>>>>>
>>>>> Any suggestions for the operation of these units and your ever expert and kind advise is always welcome. You all have a nice weekend!
>>>>>
>>>>> Kindest regards,
>>>>>
>>>>>
>>>>>
>>>>> Edgardo Molina
>>>>> Dirección IPTEL
>>>>>
>>>>> www.iptel.net.mx
>>>>>
>>>>> T : 55 55 55202444
>>>>> M : 04455 20501854
>>>>>
>>>>> Piensa en Bits SA de CV
>>>>>
>>>>>
>>>>>
>>>>> Información anexa:
>>>>>
>>>>>
>>>>>
>>>>>
>>>>> CONFIDENCIALIDAD DE INFORMACION
>>>>>
>>>>> Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
>>>>>
>>>>>
>>>>> NON-DISCLOSURE OF INFORMATION
>>>>>
>>>>> This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
>>>>>
>>>>>
>>>>>
>>>>>
>>>>>
>>>>> _______________________________________________
>>>>> time-nuts mailing list -- time-nuts@febo.com
>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>> and follow the instructions there.
>>>>>
>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts@febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
UB
Ulrich Bangert
Mon, Oct 15, 2012 5:43 AM
There's another telco model that has an MTI 260 oscillator, that I
think was manufactured in Korea but does not have the Samsung logo on
the front. I don't know much about that one.
I do own a Z3805 WITH the Samsung logo on the front AND WITH a MTI 260 inside.
Regards
Ulrich Bangert
-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Said Jackson
Gesendet: Montag, 15. Oktober 2012 00:16
An: Discussion of precise time and frequency measurement
Cc: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] HP Z3805A com port and monitoring questions
John,
I believe Edgardo bought the reprogrammed 58503A unit (was a
Z3805A before) from Ebay.
I have the same units recently purchased from yinxhu(sp?) on
Ebay and Edgardo just confirmed these are indeed 16 channel
units with true 10811 Docxos.
Explains why they work so great..
Edgardo, I use GPSCon with these, and mine are all Rs232..
Mine achieve performance in the xE-013 and noise below
-162dBc (Thunderbolt eat your heart out :). I believe there
is no better GPSDO out there presently, much less available
for sale and at those prices. PN at 1Hz is better than -105dBc..
I regularly get below 1us predicted drift as shown by the
syst:stat? command.
Bye,
Said
On Oct 14, 2012, at 10:24, John Ackermann N8UR jra@febo.com wrote:
Hi Edgardo --
I don't know a lot about the Z3805 variations other than what's on
fluke.l's site, which can be a bit confusing. All were
HP/Agilent/Symmetricom, but there were apparently at least two
"official" variations.
Mine is the Samsung version, which has the same 10811A
the Z3801, and a 16 channel Furuno GPS, 2 x 10 MHz and 2 x
dual comm ports. It's still aging in, but it seems to
the same as my older Z3801A, with stability below 1e12 from
on out.
There's another telco model that has an MTI 260 oscillator, that I
think was manufactured in Korea but does not have the
the front. I don't know much about that one.
Then there's the one that's boxed up like a 58503A bench
think is not directly from HP/Agilent/Symmetricom, but is a
repackaged by someone in China. I'm not sure what's in
the ones on eBay right now says that it has a 6 channel
was "upgraded by us which the feature is the same as 58503A". Good
luck figuring out what's in that one... (by the way, none of the
current listings are from fluke.l)
John
Edgardo Molina said the following on 10/14/2012 12:29 PM:
Dear John,
Very interesting. I haven't had the time to compare the Samsung
branded units with the typical HP units. Mine is the
has one communications port, plus 2x10 MHz and 2x 1PPS. As
am done with the late family breakfast I will do some
you know what I find.
How does the oscillators from the Samsung units compare to the HP
units? Are the Samsung units superior in any sense? Both have the
same GPS receiver/number of channels?
Thank you for opening my curiosity regarding this GPSDO model.
Best regards!
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el
destinarario de este mensaje, le suplicamos se lo notifique al
remitente mediante un correo electrónico y que borre el presente
mensaje y sus anexos de su computadora sin retener una
mismos. Queda estrictamente prohibido copiar este mensaje o hacer
usode el para cualquier propósito o divulgar su en forma parcial o
total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If
you are not the intended recipient please immediately advise the
sender by replying to this e-mail and then deleting the
its attachments from your computer without keeping a copy. It is
strictly forbidden to copy it or use it for any purpose or
its contents to any third party. Thank you.
On Oct 14, 2012, at 8:23 AM, John Ackermann N8UR jra@febo.com
wrote:
On the unit I bought from fluke.l, with a "Samsung"
two serial ports. The bottom one is the control port and is
definitely RS-232. The top port has only three wires
and I think is used for some sort of monitoring.
I've been using Ulrich's "z38xx" software to monitor it with good
success.
John
Edgardo Molina said the following on 10/13/2012 10:18 PM:
Dear Group,
Good evening. I bought two HP Z3805A units a few weeks ago. They
were bare units without antenna or power supply. After
to find a suitable 24V 3A noise free power supply for
I mounted a couple of Symmetricom antennas and just
After around 25 minutes, the GPS Lock light lighted up. As the
manual for this particular model is unobtainable in several
searches through the Internet, I found that the readily
manual for the Z3801A would cover much of the Z3905A
also got a software for monitoring the operation from BD Systems
Inc. In Colorado. Now that I am sure both are operating
only by looking at the front panel lights, I would like to ask a
couple of questions regarding their communication to the PC.
- Is the DB-25 serial port carrying RS-232 signals? I
of the Z380A1 is different: RS-422. 2. If the port is
is it possible to use a Serial to USB adapter? I have
the inconsistencies that such an adapter could cause
a TBolt. Should I assume this is the same case with the
There was once a project for building a TBolt state
line LCD screen. I own three of those monitors and they are just
great. Is there an equivalent for these HP GPS units?
Any suggestions for the operation of these units and your ever
expert and kind advise is always welcome. You all have a nice
weekend!
Kindest regards,
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el
destinarario de este mensaje, le suplicamos se lo notifique al
remitente mediante un correo electrónico y que borre el presente
mensaje y sus anexos de su computadora sin retener una
mismos. Queda estrictamente prohibido copiar este
usode el para cualquier propósito o divulgar su en forma
total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be
you are not the intended recipient please immediately advise the
sender by replying to this e-mail and then deleting the
its attachments from your computer without keeping a copy. It is
strictly forbidden to copy it or use it for any purpose
> There's another telco model that has an MTI 260 oscillator, that I
> think was manufactured in Korea but does not have the Samsung logo on
> the front. I don't know much about that one.
I do own a Z3805 WITH the Samsung logo on the front AND WITH a MTI 260 inside.
Regards
Ulrich Bangert
> -----Ursprüngliche Nachricht-----
> Von: time-nuts-bounces@febo.com
> [mailto:time-nuts-bounces@febo.com] Im Auftrag von Said Jackson
> Gesendet: Montag, 15. Oktober 2012 00:16
> An: Discussion of precise time and frequency measurement
> Cc: Discussion of precise time and frequency measurement
> Betreff: Re: [time-nuts] HP Z3805A com port and monitoring questions
>
>
> John,
>
> I believe Edgardo bought the reprogrammed 58503A unit (was a
> Z3805A before) from Ebay.
>
> I have the same units recently purchased from yinxhu(sp?) on
> Ebay and Edgardo just confirmed these are indeed 16 channel
> units with true 10811 Docxos.
>
> Explains why they work so great..
>
> Edgardo, I use GPSCon with these, and mine are all Rs232..
>
> Mine achieve performance in the xE-013 and noise below
> -162dBc (Thunderbolt eat your heart out :). I believe there
> is no better GPSDO out there presently, much less available
> for sale and at those prices. PN at 1Hz is better than -105dBc..
>
> I regularly get below 1us predicted drift as shown by the
> syst:stat? command.
>
> Bye,
> Said
>
>
>
> On Oct 14, 2012, at 10:24, John Ackermann N8UR <jra@febo.com> wrote:
>
> > Hi Edgardo --
> >
> > I don't know a lot about the Z3805 variations other than what's on
> > fluke.l's site, which can be a bit confusing. All were
> made by (for)
> > HP/Agilent/Symmetricom, but there were apparently at least two
> > "official" variations.
> >
> > Mine is the Samsung version, which has the same 10811A
> double oven as
> > the Z3801, and a 16 channel Furuno GPS, 2 x 10 MHz and 2 x
> 1 PPS, and
> > dual comm ports. It's still aging in, but it seems to
> perform about
> > the same as my older Z3801A, with stability below 1e12 from
> 1 second
> > on out.
> >
> > There's another telco model that has an MTI 260 oscillator, that I
> > think was manufactured in Korea but does not have the
> Samsung logo on
> > the front. I don't know much about that one.
> >
> > Then there's the one that's boxed up like a 58503A bench
> unit, which I
> > think is not directly from HP/Agilent/Symmetricom, but is a
> telco unit
> > repackaged by someone in China. I'm not sure what's in
> there. One of
> > the ones on eBay right now says that it has a 6 channel
> receiver, and
> > was "upgraded by us which the feature is the same as 58503A". Good
> > luck figuring out what's in that one... (by the way, none of the
> > current listings are from fluke.l)
> >
> > John
> > ----
> >
> > Edgardo Molina said the following on 10/14/2012 12:29 PM:
> >> Dear John,
> >>
> >> Very interesting. I haven't had the time to compare the Samsung
> >> branded units with the typical HP units. Mine is the
> later. It only
> >> has one communications port, plus 2x10 MHz and 2x 1PPS. As
> soon as I
> >> am done with the late family breakfast I will do some
> testing and let
> >> you know what I find.
> >>
> >> How does the oscillators from the Samsung units compare to the HP
> >> units? Are the Samsung units superior in any sense? Both have the
> >> same GPS receiver/number of channels?
> >>
> >> Thank you for opening my curiosity regarding this GPSDO model.
> >>
> >> Best regards!
> >>
> >>
> >>
> >> Edgardo Molina
> >> Dirección IPTEL
> >>
> >> www.iptel.net.mx
> >>
> >> T : 55 55 55202444
> >> M : 04455 20501854
> >>
> >> Piensa en Bits SA de CV
> >>
> >>
> >>
> >> Información anexa:
> >>
> >>
> >>
> >>
> >> CONFIDENCIALIDAD DE INFORMACION
> >>
> >> Este mensaje tiene carácter confidencial. Si usted no es el
> >> destinarario de este mensaje, le suplicamos se lo notifique al
> >> remitente mediante un correo electrónico y que borre el presente
> >> mensaje y sus anexos de su computadora sin retener una
> copia de los
> >> mismos. Queda estrictamente prohibido copiar este mensaje o hacer
> >> usode el para cualquier propósito o divulgar su en forma parcial o
> >> total su contenido. Gracias.
> >>
> >>
> >> NON-DISCLOSURE OF INFORMATION
> >>
> >> This email is strictly confidential and may also be privileged. If
> >> you are not the intended recipient please immediately advise the
> >> sender by replying to this e-mail and then deleting the
> message and
> >> its attachments from your computer without keeping a copy. It is
> >> strictly forbidden to copy it or use it for any purpose or
> disclose
> >> its contents to any third party. Thank you.
> >>
> >>
> >>
> >>
> >>
> >> On Oct 14, 2012, at 8:23 AM, John Ackermann N8UR <jra@febo.com>
> >> wrote:
> >>
> >>> On the unit I bought from fluke.l, with a "Samsung"
> label, there are
> >>> two serial ports. The bottom one is the control port and is
> >>> definitely RS-232. The top port has only three wires
> (TXD, RXD, SG)
> >>> and I think is used for some sort of monitoring.
> >>>
> >>> I've been using Ulrich's "z38xx" software to monitor it with good
> >>> success.
> >>>
> >>> John
> >>> ----
> >>>
> >>> Edgardo Molina said the following on 10/13/2012 10:18 PM:
> >>>> Dear Group,
> >>>>
> >>>> Good evening. I bought two HP Z3805A units a few weeks ago. They
> >>>> were bare units without antenna or power supply. After
> struggling
> >>>> to find a suitable 24V 3A noise free power supply for
> each of them,
> >>>> I mounted a couple of Symmetricom antennas and just
> fired them up.
> >>>>
> >>>> After around 25 minutes, the GPS Lock light lighted up. As the
> >>>> manual for this particular model is unobtainable in several
> >>>> searches through the Internet, I found that the readily
> available
> >>>> manual for the Z3801A would cover much of the Z3905A
> operation. I
> >>>> also got a software for monitoring the operation from BD Systems
> >>>> Inc. In Colorado. Now that I am sure both are operating
> and locked
> >>>> only by looking at the front panel lights, I would like to ask a
> >>>> couple of questions regarding their communication to the PC.
> >>>>
> >>>> 1. Is the DB-25 serial port carrying RS-232 signals? I
> see the port
> >>>> of the Z380A1 is different: RS-422. 2. If the port is
> plain RS-232
> >>>> is it possible to use a Serial to USB adapter? I have
> heard about
> >>>> the inconsistencies that such an adapter could cause
> when used with
> >>>> a TBolt. Should I assume this is the same case with the
> Z3805A? 3.
> >>>> There was once a project for building a TBolt state
> monitor on a 2
> >>>> line LCD screen. I own three of those monitors and they are just
> >>>> great. Is there an equivalent for these HP GPS units?
> >>>>
> >>>> Any suggestions for the operation of these units and your ever
> >>>> expert and kind advise is always welcome. You all have a nice
> >>>> weekend!
> >>>>
> >>>> Kindest regards,
> >>>>
> >>>>
> >>>>
> >>>> Edgardo Molina
> >>>> Dirección IPTEL
> >>>>
> >>>> www.iptel.net.mx
> >>>>
> >>>> T : 55 55 55202444
> >>>> M : 04455 20501854
> >>>>
> >>>> Piensa en Bits SA de CV
> >>>>
> >>>>
> >>>>
> >>>> Información anexa:
> >>>>
> >>>>
> >>>>
> >>>>
> >>>> CONFIDENCIALIDAD DE INFORMACION
> >>>>
> >>>> Este mensaje tiene carácter confidencial. Si usted no es el
> >>>> destinarario de este mensaje, le suplicamos se lo notifique al
> >>>> remitente mediante un correo electrónico y que borre el presente
> >>>> mensaje y sus anexos de su computadora sin retener una
> copia de los
> >>>> mismos. Queda estrictamente prohibido copiar este
> mensaje o hacer
> >>>> usode el para cualquier propósito o divulgar su en forma
> parcial o
> >>>> total su contenido. Gracias.
> >>>>
> >>>>
> >>>> NON-DISCLOSURE OF INFORMATION
> >>>>
> >>>> This email is strictly confidential and may also be
> privileged. If
> >>>> you are not the intended recipient please immediately advise the
> >>>> sender by replying to this e-mail and then deleting the
> message and
> >>>> its attachments from your computer without keeping a copy. It is
> >>>> strictly forbidden to copy it or use it for any purpose
> or disclose
> >>>> its contents to any third party. Thank you.
> >>>>
> >>>>
> >>>>
> >>>>
> >>>>
> >>>> _______________________________________________
> >>>> time-nuts mailing list -- time-nuts@febo.com
> >>>> To unsubscribe, go to
> >>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> >>>> and follow the instructions there.
> >>>>
> >>
> >> _______________________________________________
> >> time-nuts mailing list -- time-nuts@febo.com
> >> To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> >> and follow the instructions there.
> >>
> >
> > _______________________________________________
> > time-nuts mailing list -- time-nuts@febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> > and follow the instructions there.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
R
Rex
Mon, Oct 15, 2012 10:11 AM
On 10/14/2012 10:43 PM, Ulrich Bangert wrote:
There's another telco model that has an MTI 260 oscillator, that I
think was manufactured in Korea but does not have the Samsung logo on
the front. I don't know much about that one.
I do own a Z3805 WITH the Samsung logo on the front AND WITH a MTI 260 inside.
Regards
Ulrich Bangert
Me too -- same description of Z3805. For some reason the output of that
3805 seems a bit cleaner (in PN) than my Z3816A which also has an MTI
260 oscillator.
BTW, Ulrich, I use your software as the main monitor for both. Thanks
for all your work!
On 10/14/2012 10:43 PM, Ulrich Bangert wrote:
>> There's another telco model that has an MTI 260 oscillator, that I
>> think was manufactured in Korea but does not have the Samsung logo on
>> the front. I don't know much about that one.
> I do own a Z3805 WITH the Samsung logo on the front AND WITH a MTI 260 inside.
>
> Regards
> Ulrich Bangert
>
Me too -- same description of Z3805. For some reason the output of that
3805 seems a bit cleaner (in PN) than my Z3816A which also has an MTI
260 oscillator.
BTW, Ulrich, I use your software as the main monitor for both. Thanks
for all your work!
DI
David I. Emery
Thu, Oct 18, 2012 3:33 AM
On Sun, Oct 14, 2012 at 03:15:53PM -0700, Said Jackson wrote:
I believe Edgardo bought the reprogrammed 58503A unit (was a Z3805A
before) from Ebay.
I have the same units recently purchased from yinxhu(sp?) on Ebay and
Edgardo just confirmed these are indeed 16 channel units with true 10811
Docxos.
I just bought one from yinuxhk based on your comments last
weekend and it arrived today.
Has one RS-232 port and two 10 MHz and two PPS. XLR 24 volt
power connector. Made in Korea, marked HP Z3805A std.
In very good shape. Seems like almost new old stock. HP DOCXO.
Seems rather old based on serial number, serial is 3625Axxxxx
which would be 1996 if IIRC.
I'm pretty sure what Edgardo bought exactly.
Are yours this exact version and what age are they (from the
serial number) ?
--
Dave Emery N1PRE/AE, die@dieconsulting.com DIE Consulting, Weston, Mass 02493
"An empty zombie mind with a forlorn barely readable weatherbeaten
'For Rent' sign still vainly flapping outside on the weed encrusted pole - in
celebration of what could have been, but wasn't and is not to be now either."
On Sun, Oct 14, 2012 at 03:15:53PM -0700, Said Jackson wrote:
>
> I believe Edgardo bought the reprogrammed 58503A unit (was a Z3805A
> before) from Ebay.
>
> I have the same units recently purchased from yinxhu(sp?) on Ebay and
> Edgardo just confirmed these are indeed 16 channel units with true 10811
> Docxos.
I just bought one from yinuxhk based on your comments last
weekend and it arrived today.
Has one RS-232 port and two 10 MHz and two PPS. XLR 24 volt
power connector. Made in Korea, marked HP Z3805A std.
In very good shape. Seems like almost new old stock. HP DOCXO.
Seems rather old based on serial number, serial is 3625Axxxxx
which would be 1996 if IIRC.
I'm pretty sure what Edgardo bought exactly.
Are yours this exact version and what age are they (from the
serial number) ?
--
Dave Emery N1PRE/AE, die@dieconsulting.com DIE Consulting, Weston, Mass 02493
"An empty zombie mind with a forlorn barely readable weatherbeaten
'For Rent' sign still vainly flapping outside on the weed encrusted pole - in
celebration of what could have been, but wasn't and is not to be now either."
EM
Edgardo Molina
Thu, Oct 18, 2012 5:47 AM
Dear David,
Yes, I bought the same unit as you. I will come back to you tomorrow with some serial number information.
I am starting to think my serial cable is bad. I can't connect to any of my two units! Both of them have been locked to GPS signals perfectly. Shame on me… I just returned from CENAM and I am very, very tired. If you are able to connect to your unit, please advise. None of the combinations I got in past emails worked. It must be cable or serial port, or blame it on the operator. I will update this thread with more information tomorrow.
I will be getting new toys to play around during the weekend. I will comment in a separate thread as soon as they are delivered to me. I wish to correspond with someone having the same equipment nowadays ; )
Thanks.
Best regards and good night!
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 17, 2012, at 10:33 PM, "David I. Emery" die@dieconsulting.com wrote:
On Sun, Oct 14, 2012 at 03:15:53PM -0700, Said Jackson wrote:
I believe Edgardo bought the reprogrammed 58503A unit (was a Z3805A
before) from Ebay.
I have the same units recently purchased from yinxhu(sp?) on Ebay and
Edgardo just confirmed these are indeed 16 channel units with true 10811
Docxos.
I just bought one from yinuxhk based on your comments last
weekend and it arrived today.
Has one RS-232 port and two 10 MHz and two PPS. XLR 24 volt
power connector. Made in Korea, marked HP Z3805A std.
In very good shape. Seems like almost new old stock. HP DOCXO.
Seems rather old based on serial number, serial is 3625Axxxxx
which would be 1996 if IIRC.
I'm pretty sure what Edgardo bought exactly.
Are yours this exact version and what age are they (from the
serial number) ?
--
Dave Emery N1PRE/AE, die@dieconsulting.com DIE Consulting, Weston, Mass 02493
"An empty zombie mind with a forlorn barely readable weatherbeaten
'For Rent' sign still vainly flapping outside on the weed encrusted pole - in
celebration of what could have been, but wasn't and is not to be now either."
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Dear David,
Yes, I bought the same unit as you. I will come back to you tomorrow with some serial number information.
I am starting to think my serial cable is bad. I can't connect to any of my two units! Both of them have been locked to GPS signals perfectly. Shame on me… I just returned from CENAM and I am very, very tired. If you are able to connect to your unit, please advise. None of the combinations I got in past emails worked. It must be cable or serial port, or blame it on the operator. I will update this thread with more information tomorrow.
I will be getting new toys to play around during the weekend. I will comment in a separate thread as soon as they are delivered to me. I wish to correspond with someone having the same equipment nowadays ; )
Thanks.
Best regards and good night!
Edgardo Molina
Dirección IPTEL
www.iptel.net.mx
T : 55 55 55202444
M : 04455 20501854
Piensa en Bits SA de CV
Información anexa:
CONFIDENCIALIDAD DE INFORMACION
Este mensaje tiene carácter confidencial. Si usted no es el destinarario de este mensaje, le suplicamos se lo notifique al remitente mediante un correo electrónico y que borre el presente mensaje y sus anexos de su computadora sin retener una copia de los mismos. Queda estrictamente prohibido copiar este mensaje o hacer usode el para cualquier propósito o divulgar su en forma parcial o total su contenido. Gracias.
NON-DISCLOSURE OF INFORMATION
This email is strictly confidential and may also be privileged. If you are not the intended recipient please immediately advise the sender by replying to this e-mail and then deleting the message and its attachments from your computer without keeping a copy. It is strictly forbidden to copy it or use it for any purpose or disclose its contents to any third party. Thank you.
On Oct 17, 2012, at 10:33 PM, "David I. Emery" <die@dieconsulting.com> wrote:
> On Sun, Oct 14, 2012 at 03:15:53PM -0700, Said Jackson wrote:
>>
>
>> I believe Edgardo bought the reprogrammed 58503A unit (was a Z3805A
>> before) from Ebay.
>>
>
>> I have the same units recently purchased from yinxhu(sp?) on Ebay and
>> Edgardo just confirmed these are indeed 16 channel units with true 10811
>> Docxos.
>
> I just bought one from yinuxhk based on your comments last
> weekend and it arrived today.
>
> Has one RS-232 port and two 10 MHz and two PPS. XLR 24 volt
> power connector. Made in Korea, marked HP Z3805A std.
>
> In very good shape. Seems like almost new old stock. HP DOCXO.
>
> Seems rather old based on serial number, serial is 3625Axxxxx
> which would be 1996 if IIRC.
>
> I'm pretty sure what Edgardo bought exactly.
>
> Are yours this exact version and what age are they (from the
> serial number) ?
>
>
>
> --
> Dave Emery N1PRE/AE, die@dieconsulting.com DIE Consulting, Weston, Mass 02493
> "An empty zombie mind with a forlorn barely readable weatherbeaten
> 'For Rent' sign still vainly flapping outside on the weed encrusted pole - in
> celebration of what could have been, but wasn't and is not to be now either."
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
DI
David I. Emery
Sat, Oct 20, 2012 6:19 AM
On Thu, Oct 18, 2012 at 12:47:30AM -0500, Edgardo Molina wrote:
Yes, I bought the same unit as you. I will come back to you tomorrow
with some serial number information.
I found mine is a DTE with TXD on pin 2... RXD on pin 3.
Seems to like 9600,8,n,1
Have it working under a XP virtual machine on this laptop with
the K8CU GPS receiver control program. Seems to work OK...
no stats on performance yet.
Mine has a Motorola 6 channel GPS rx in it - not a Furuno 16
channel.
In the same shipment I also ordered one of the Samsung branded
Symmetricom GPSes that does have the Furuno 16 channel receiver.
K8CU GPSCON reports it has a GT-80 and shows 16 channels. It has
a Symmetricom OCXO, looks MTIish... but not MTI. Mounted on a
doubler board - outputs 10 MHz. Looking inside the receiver
is a Furuno GT8031F.
I'm sort of wondering if I could upgrade the GPS in my close to
NOS HP z3805 box (would obviously require using the Samsung version of
the firmware) from the Samsung box... not clear how much (other than
TRIAM and WAAS) that might buy me or whether it would be worth it.
Though I am sure you know all this, you certainly should be able
to find the port pins with a DVM or scope - mine outputs about -8 volts
on TXD and RXD floats about 0.2 volts positive. I did find I need a
null modem cable to connect it up to a regular serial port (for this
test a keyspan HS-19 USB serial port).
I'm not exactly sure what Said Jackson has and sings the praises
of. Fluke.l has been selling a few Samsungs and has one advertised that
I guess is similar to mine but with an explicit MTI 260.
I presume Said has units with the 10811-60165 rather than the
MTI-260... I'm curious which oscillator he is finding those stability
and phase noise specs on - I'm looking to upgrade the reference to some
microwave LO chains and would like a nice box with that performance.
I'm not sure when he bought his or what age they are.
Apparently some of the HP branked z3805s must have had the Furuno
receiver module.
There is also tommy_chou who has some Samsung z8505As with
the 10811 DOCXO... not sure what GPS these have.
Appear to be 3 sellers with similar ex-Korea CDMA stuff...
tommy_chou, our familiar fluke.1 and yixunhk.
you are able to connect to your unit, please advise. None of the
combinations I got in past emails worked. It must be cable or serial
port, or blame it on the operator. I will update this thread with more
information tomorrow.
I have found the obvious pilot error issues with things like
hyperterminal - the box should respond to a *CLS command with scpi > and
to an *IDN? command with the serial number. I did find (silly me) that
hyperterminal defaults to DCD flow control on and you have to say no
flow control before it will see anything from the ports on the box. This
is because DCD comes out off with the Null modem cable I was using.
Basic stuff.
I will be getting new toys to play around during the weekend. I will
comment in a separate thread as soon as they are delivered to me. I wish
to correspond with someone having the same equipment nowadays ; )
--
Dave Emery N1PRE/AE, die@dieconsulting.com DIE Consulting, Weston, Mass 02493
"An empty zombie mind with a forlorn barely readable weatherbeaten
'For Rent' sign still vainly flapping outside on the weed encrusted pole - in
celebration of what could have been, but wasn't and is not to be now either."
On Thu, Oct 18, 2012 at 12:47:30AM -0500, Edgardo Molina wrote:
> Dear David,
>
> Yes, I bought the same unit as you. I will come back to you tomorrow
> with some serial number information.
I found mine is a DTE with TXD on pin 2... RXD on pin 3.
Seems to like 9600,8,n,1
Have it working under a XP virtual machine on this laptop with
the K8CU GPS receiver control program. Seems to work OK...
no stats on performance yet.
Mine has a Motorola 6 channel GPS rx in it - not a Furuno 16
channel.
In the same shipment I also ordered one of the Samsung branded
Symmetricom GPSes that does have the Furuno 16 channel receiver.
K8CU GPSCON reports it has a GT-80 and shows 16 channels. It has
a Symmetricom OCXO, looks MTIish... but not MTI. Mounted on a
doubler board - outputs 10 MHz. Looking inside the receiver
is a Furuno GT8031F.
I'm sort of wondering if I could upgrade the GPS in my close to
NOS HP z3805 box (would obviously require using the Samsung version of
the firmware) from the Samsung box... not clear how much (other than
TRIAM and WAAS) that might buy me or whether it would be worth it.
Though I am sure you know all this, you certainly should be able
to find the port pins with a DVM or scope - mine outputs about -8 volts
on TXD and RXD floats about 0.2 volts positive. I did find I need a
null modem cable to connect it up to a regular serial port (for this
test a keyspan HS-19 USB serial port).
I'm not exactly sure what Said Jackson has and sings the praises
of. Fluke.l has been selling a few Samsungs and has one advertised that
I guess is similar to mine but with an explicit MTI 260.
I presume Said has units with the 10811-60165 rather than the
MTI-260... I'm curious which oscillator he is finding those stability
and phase noise specs on - I'm looking to upgrade the reference to some
microwave LO chains and would like a nice box with that performance.
I'm not sure when he bought his or what age they are.
Apparently some of the HP branked z3805s must have had the Furuno
receiver module.
There is also tommy_chou who has some Samsung z8505As with
the 10811 DOCXO... not sure what GPS these have.
Appear to be 3 sellers with similar ex-Korea CDMA stuff...
tommy_chou, our familiar fluke.1 and yixunhk.
> you are able to connect to your unit, please advise. None of the
> combinations I got in past emails worked. It must be cable or serial
> port, or blame it on the operator. I will update this thread with more
> information tomorrow.
I have found the obvious pilot error issues with things like
hyperterminal - the box should respond to a *CLS command with scpi > and
to an *IDN? command with the serial number. I did find (silly me) that
hyperterminal defaults to DCD flow control on and you have to say no
flow control before it will see anything from the ports on the box. This
is because DCD comes out off with the Null modem cable I was using.
Basic stuff.
> I will be getting new toys to play around during the weekend. I will
> comment in a separate thread as soon as they are delivered to me. I wish
> to correspond with someone having the same equipment nowadays ; )
--
Dave Emery N1PRE/AE, die@dieconsulting.com DIE Consulting, Weston, Mass 02493
"An empty zombie mind with a forlorn barely readable weatherbeaten
'For Rent' sign still vainly flapping outside on the weed encrusted pole - in
celebration of what could have been, but wasn't and is not to be now either."
TV
Tom Van Baak
Mon, Oct 22, 2012 5:13 AM
Trimble and HP took different approaches on their GPSDO's.
Trimble lets you get into things and fiddle around. HP is much more a black box.
I have never seen anything that lets you "fiddle' with one of the HP units.
Hi Bob,
I'm glad you asked. True, the Trimble TSIP binary command set lets you fiddle and the HP SCPI command set is much easier to use but more limited by comparison. But there's also a rich under-the-hood layer on the HP units. Spend some time with this file and you'll see what I mean:
http://leapsecond.com/museum/z3801a/z3801a-bin.txt
I explored the functionality in some detail during my hp GPSDO DAC dither investigation some years ago. Contact me offline.
/tvb
> Trimble and HP took different approaches on their GPSDO's.
> Trimble lets you get into things and fiddle around. HP is much more a black box.
> I have never seen anything that lets you "fiddle' with one of the HP units.
Hi Bob,
I'm glad you asked. True, the Trimble TSIP binary command set lets you fiddle and the HP SCPI command set is much easier to use but more limited by comparison. But there's also a rich under-the-hood layer on the HP units. Spend some time with this file and you'll see what I mean:
http://leapsecond.com/museum/z3801a/z3801a-bin.txt
I explored the functionality in some detail during my hp GPSDO DAC dither investigation some years ago. Contact me offline.
/tvb
DH
David Hooke
Mon, Oct 22, 2012 6:07 AM
Tom,
It looks like they have a pForth interpreter running in there under
PSOS: http://code.google.com/p/pforth/. Is there a mechanism to switch
from SCPI to the Forth interpreter?
david
Trimble and HP took different approaches on their GPSDO's.
Trimble lets you get into things and fiddle around. HP is much more a black box.
I have never seen anything that lets you "fiddle' with one of the HP units.
Hi Bob,
I'm glad you asked. True, the Trimble TSIP binary command set lets you fiddle and the HP SCPI command set is much easier to use but more limited by comparison. But there's also a rich under-the-hood layer on the HP units. Spend some time with this file and you'll see what I mean:
http://leapsecond.com/museum/z3801a/z3801a-bin.txt
I explored the functionality in some detail during my hp GPSDO DAC dither investigation some years ago. Contact me offline.
/tvb
Tom,
It looks like they have a pForth interpreter running in there under
PSOS: <http://code.google.com/p/pforth/>. Is there a mechanism to switch
from SCPI to the Forth interpreter?
david
>> Trimble and HP took different approaches on their GPSDO's.
>> Trimble lets you get into things and fiddle around. HP is much more a black box.
>> I have never seen anything that lets you "fiddle' with one of the HP units.
> Hi Bob,
>
> I'm glad you asked. True, the Trimble TSIP binary command set lets you fiddle and the HP SCPI command set is much easier to use but more limited by comparison. But there's also a rich under-the-hood layer on the HP units. Spend some time with this file and you'll see what I mean:
> http://leapsecond.com/museum/z3801a/z3801a-bin.txt
> I explored the functionality in some detail during my hp GPSDO DAC dither investigation some years ago. Contact me offline.
>
> /tvb
>
>