time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

FS740 Thoughts?

MH
Matt Huszagh
Wed, Nov 17, 2021 5:50 AM

Hi,

I'm curious to hear people's thoughts on the SRS FS740
(https://www.thinksrs.com/products/fs740.html) as a lab frequency
standard. The phase noise performance of the OCXO and Rb options seem
quite good (the same specs apparently apply during GPS disciplining too)

| offset |  TCXO  | OCXO/Rb  |
| (Hz)  | (dBc/Hz) | (dBc/Hz) |
|--------+----------+----------|
| 1      |  -90    |  -108  |
| 10    |  -104  |  -130  |
| 100    |  -125  |  -144  |
| 1k    |  -142  |  -143  |
| 10k    |  -147  |  -148  |
| 100k  |  -149  |  -149  |
| 1M    |  -153  |  -153  |

as does the Allan deviation (during GPS disciplining)

| time interval (s) | TCXO  | OCXO  |  Rb  |
|-------------------+-------+-------+-------|
| 0.01              | 1E-10 | 2E-11 | 2E-11 |
| 0.1              | 4E-11 | 4E-12 | 4E-12 |
| 1                | 3E-11 | 2E-12 | 1E-11 |
| 10                | 1E-10 | 2E-12 | 6E-12 |
| 100              | 6E-11 | 2E-12 | 1E-12 |
| 1k                | 6E-12 | 3E-12 | 1E-12 |
| 10k              | 6E-13 | 6E-13 | 3E-13 |
| 100k              | 6E-14 | 6E-14 | 3E-14 |
| 1M                | 5E-15 | 5E-15 | 3E-15 |

(I used the typical performance plots for the phase noise and Allan
deviation data)

The OCXO does not exhibit the the hump around 2s seen by the rubidium
oscillator option (PRS10) and therefore could be seen as better in
applications not requiring the better holdover performance by the Rb.

It also provides a pulse output with jitter < 50 ps RMS that could be
used as a PPS signal. It can also provide a low noise 100 MHz output
signal and can act as a frequency counter and DDS with high resolution
and accuracy. And it comes with the schematics! But.... it's definitely
not cheap. Other thoughts/opinions?

Matt

Hi, I'm curious to hear people's thoughts on the SRS FS740 (https://www.thinksrs.com/products/fs740.html) as a lab frequency standard. The phase noise performance of the OCXO and Rb options seem quite good (the same specs apparently apply during GPS disciplining too) | offset | TCXO | OCXO/Rb | | (Hz) | (dBc/Hz) | (dBc/Hz) | |--------+----------+----------| | 1 | -90 | -108 | | 10 | -104 | -130 | | 100 | -125 | -144 | | 1k | -142 | -143 | | 10k | -147 | -148 | | 100k | -149 | -149 | | 1M | -153 | -153 | as does the Allan deviation (during GPS disciplining) | time interval (s) | TCXO | OCXO | Rb | |-------------------+-------+-------+-------| | 0.01 | 1E-10 | 2E-11 | 2E-11 | | 0.1 | 4E-11 | 4E-12 | 4E-12 | | 1 | 3E-11 | 2E-12 | 1E-11 | | 10 | 1E-10 | 2E-12 | 6E-12 | | 100 | 6E-11 | 2E-12 | 1E-12 | | 1k | 6E-12 | 3E-12 | 1E-12 | | 10k | 6E-13 | 6E-13 | 3E-13 | | 100k | 6E-14 | 6E-14 | 3E-14 | | 1M | 5E-15 | 5E-15 | 3E-15 | (I used the typical performance plots for the phase noise and Allan deviation data) The OCXO does not exhibit the the hump around 2s seen by the rubidium oscillator option (PRS10) and therefore could be seen as better in applications not requiring the better holdover performance by the Rb. It also provides a pulse output with jitter < 50 ps RMS that could be used as a PPS signal. It can also provide a low noise 100 MHz output signal and can act as a frequency counter and DDS with high resolution and accuracy. And it comes with the schematics! But.... it's definitely not cheap. Other thoughts/opinions? Matt
BK
Bob kb8tq
Wed, Nov 17, 2021 1:56 PM

Hi

I have a 740 as do others on the list. The big gotcha is the OCXO in the device.
It has various quirks that limit the performance of the device. The ADEV plot they
show for the OCXO is a “sometimes / maybe / might” sort of thing. One of the
“get to it eventually” projects here is to see if it can be replaced with something
that has better ADEV performance.

Be wary of folks showing up in dirt parking lots in the middle of Montana with one
in the back of their car …. :) Doubly so if the car shows signs of having hauled
a dozen Cs standards up various mountains….. If they are named Tom, simply
give up and accept that once you see one you will eventually buy one…

The process goes sort of like: Gee, that’s a lot of money. Later on, … I need a few
more standard lines at this or that frequency … Then it goes to If only I could come
up with that signal … hmmm … 740 would do that … hmmmm…. off you go to
get one.

It’s a handy box to do a variety of things with. The only issue I’ve seen is the OCXO
performance.

Bob

On Nov 17, 2021, at 12:50 AM, Matt Huszagh huszaghmatt@gmail.com wrote:

Hi,

I'm curious to hear people's thoughts on the SRS FS740
(https://www.thinksrs.com/products/fs740.html) as a lab frequency
standard. The phase noise performance of the OCXO and Rb options seem
quite good (the same specs apparently apply during GPS disciplining too)

| offset |  TCXO  | OCXO/Rb  |
| (Hz)  | (dBc/Hz) | (dBc/Hz) |
|--------+----------+----------|
| 1      |  -90    |  -108  |
| 10    |  -104  |  -130  |
| 100    |  -125  |  -144  |
| 1k    |  -142  |  -143  |
| 10k    |  -147  |  -148  |
| 100k  |  -149  |  -149  |
| 1M    |  -153  |  -153  |

as does the Allan deviation (during GPS disciplining)

| time interval (s) | TCXO  | OCXO  |  Rb  |
|-------------------+-------+-------+-------|
| 0.01              | 1E-10 | 2E-11 | 2E-11 |
| 0.1              | 4E-11 | 4E-12 | 4E-12 |
| 1                | 3E-11 | 2E-12 | 1E-11 |
| 10                | 1E-10 | 2E-12 | 6E-12 |
| 100              | 6E-11 | 2E-12 | 1E-12 |
| 1k                | 6E-12 | 3E-12 | 1E-12 |
| 10k              | 6E-13 | 6E-13 | 3E-13 |
| 100k              | 6E-14 | 6E-14 | 3E-14 |
| 1M                | 5E-15 | 5E-15 | 3E-15 |

(I used the typical performance plots for the phase noise and Allan
deviation data)

The OCXO does not exhibit the the hump around 2s seen by the rubidium
oscillator option (PRS10) and therefore could be seen as better in
applications not requiring the better holdover performance by the Rb.

It also provides a pulse output with jitter < 50 ps RMS that could be
used as a PPS signal. It can also provide a low noise 100 MHz output
signal and can act as a frequency counter and DDS with high resolution
and accuracy. And it comes with the schematics! But.... it's definitely
not cheap. Other thoughts/opinions?

Matt


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

Hi I have a 740 as do others on the list. The big gotcha is the OCXO in the device. It has various quirks that limit the performance of the device. The ADEV plot they show for the OCXO is a “sometimes / maybe / might” sort of thing. One of the “get to it eventually” projects here is to see if it can be replaced with something that has better ADEV performance. Be wary of folks showing up in dirt parking lots in the middle of Montana with one in the back of their car …. :) Doubly so if the car shows signs of having hauled a dozen Cs standards up various mountains….. If they are named Tom, simply give up and accept that once you see one you will eventually buy one… The process goes sort of like: Gee, that’s a lot of money. Later on, … I need a few more standard lines at this or that frequency … Then it goes to If only I could come up with *that* signal … hmmm … 740 would do that … hmmmm…. off you go to get one. It’s a handy box to do a variety of things with. The only issue I’ve seen is the OCXO performance. Bob > On Nov 17, 2021, at 12:50 AM, Matt Huszagh <huszaghmatt@gmail.com> wrote: > > Hi, > > I'm curious to hear people's thoughts on the SRS FS740 > (https://www.thinksrs.com/products/fs740.html) as a lab frequency > standard. The phase noise performance of the OCXO and Rb options seem > quite good (the same specs apparently apply during GPS disciplining too) > > | offset | TCXO | OCXO/Rb | > | (Hz) | (dBc/Hz) | (dBc/Hz) | > |--------+----------+----------| > | 1 | -90 | -108 | > | 10 | -104 | -130 | > | 100 | -125 | -144 | > | 1k | -142 | -143 | > | 10k | -147 | -148 | > | 100k | -149 | -149 | > | 1M | -153 | -153 | > > as does the Allan deviation (during GPS disciplining) > > | time interval (s) | TCXO | OCXO | Rb | > |-------------------+-------+-------+-------| > | 0.01 | 1E-10 | 2E-11 | 2E-11 | > | 0.1 | 4E-11 | 4E-12 | 4E-12 | > | 1 | 3E-11 | 2E-12 | 1E-11 | > | 10 | 1E-10 | 2E-12 | 6E-12 | > | 100 | 6E-11 | 2E-12 | 1E-12 | > | 1k | 6E-12 | 3E-12 | 1E-12 | > | 10k | 6E-13 | 6E-13 | 3E-13 | > | 100k | 6E-14 | 6E-14 | 3E-14 | > | 1M | 5E-15 | 5E-15 | 3E-15 | > > (I used the typical performance plots for the phase noise and Allan > deviation data) > > The OCXO does not exhibit the the hump around 2s seen by the rubidium > oscillator option (PRS10) and therefore could be seen as better in > applications not requiring the better holdover performance by the Rb. > > It also provides a pulse output with jitter < 50 ps RMS that could be > used as a PPS signal. It can also provide a low noise 100 MHz output > signal and can act as a frequency counter and DDS with high resolution > and accuracy. And it comes with the schematics! But.... it's definitely > not cheap. Other thoughts/opinions? > > Matt > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there.
DW
Dana Whitlow
Wed, Nov 17, 2021 2:08 PM

I think that before one places much faith in the ADEV data, it's important
to understand the thermal environment in which the tests were conducted,
compared to your own thermal environment.  The tests were likely to have
been performed in a well-stabilized ambient temperature, to make them
look good.  But your own results may be a lot worse due to HVAC cycling
etc.

One way to reduce such effects is to make the GPS locking loop quite fast,
but doing so is at the cost of the disciplines oscillator's being yanked
around
in phase by "GPS noise", which is really quite awful.  At the other
extreme,
(long locking loop time constant) one largely escapes GPS noise, but finds
that his system has essentially become an expensive thermometer.  So I favor
the use of a GPS-disciplined Rb for critical work (or an active H-maser if
you're
really rich).

I'm presently crutching along with a manually-disciplined surplus PRS-10,
but
am looking for a good excuse to get a factory new lockable unit.

Aha:  KB8TQ's note just came in;  we're saying about the same thing- I'm
just
wordier.

Dana  K8YUM

On Tue, Nov 16, 2021 at 11:59 PM Matt Huszagh huszaghmatt@gmail.com wrote:

Hi,

I'm curious to hear people's thoughts on the SRS FS740
(https://www.thinksrs.com/products/fs740.html) as a lab frequency
standard. The phase noise performance of the OCXO and Rb options seem
quite good (the same specs apparently apply during GPS disciplining too)

| offset |  TCXO  | OCXO/Rb  |
| (Hz)  | (dBc/Hz) | (dBc/Hz) |
|--------+----------+----------|
| 1      |  -90    |  -108  |
| 10    |  -104  |  -130  |
| 100    |  -125  |  -144  |
| 1k    |  -142  |  -143  |
| 10k    |  -147  |  -148  |
| 100k  |  -149  |  -149  |
| 1M    |  -153  |  -153  |

as does the Allan deviation (during GPS disciplining)

| time interval (s) | TCXO  | OCXO  |  Rb  |
|-------------------+-------+-------+-------|
| 0.01              | 1E-10 | 2E-11 | 2E-11 |
| 0.1              | 4E-11 | 4E-12 | 4E-12 |
| 1                | 3E-11 | 2E-12 | 1E-11 |
| 10                | 1E-10 | 2E-12 | 6E-12 |
| 100              | 6E-11 | 2E-12 | 1E-12 |
| 1k                | 6E-12 | 3E-12 | 1E-12 |
| 10k              | 6E-13 | 6E-13 | 3E-13 |
| 100k              | 6E-14 | 6E-14 | 3E-14 |
| 1M                | 5E-15 | 5E-15 | 3E-15 |

(I used the typical performance plots for the phase noise and Allan
deviation data)

The OCXO does not exhibit the the hump around 2s seen by the rubidium
oscillator option (PRS10) and therefore could be seen as better in
applications not requiring the better holdover performance by the Rb.

It also provides a pulse output with jitter < 50 ps RMS that could be
used as a PPS signal. It can also provide a low noise 100 MHz output
signal and can act as a frequency counter and DDS with high resolution
and accuracy. And it comes with the schematics! But.... it's definitely
not cheap. Other thoughts/opinions?

Matt


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

I think that before one places much faith in the ADEV data, it's important to understand the thermal environment in which the tests were conducted, compared to your own thermal environment. The tests were likely to have been performed in a well-stabilized ambient temperature, to make them look good. But your own results may be a lot worse due to HVAC cycling etc. One way to reduce such effects is to make the GPS locking loop quite fast, but doing so is at the cost of the disciplines oscillator's being yanked around in phase by "GPS noise", which is really quite awful. At the other extreme, (long locking loop time constant) one largely escapes GPS noise, but finds that his system has essentially become an expensive thermometer. So I favor the use of a GPS-disciplined Rb for critical work (or an active H-maser if you're really rich). I'm presently crutching along with a manually-disciplined surplus PRS-10, but am looking for a good excuse to get a factory new lockable unit. Aha: KB8TQ's note just came in; we're saying about the same thing- I'm just wordier. Dana K8YUM On Tue, Nov 16, 2021 at 11:59 PM Matt Huszagh <huszaghmatt@gmail.com> wrote: > Hi, > > I'm curious to hear people's thoughts on the SRS FS740 > (https://www.thinksrs.com/products/fs740.html) as a lab frequency > standard. The phase noise performance of the OCXO and Rb options seem > quite good (the same specs apparently apply during GPS disciplining too) > > | offset | TCXO | OCXO/Rb | > | (Hz) | (dBc/Hz) | (dBc/Hz) | > |--------+----------+----------| > | 1 | -90 | -108 | > | 10 | -104 | -130 | > | 100 | -125 | -144 | > | 1k | -142 | -143 | > | 10k | -147 | -148 | > | 100k | -149 | -149 | > | 1M | -153 | -153 | > > as does the Allan deviation (during GPS disciplining) > > | time interval (s) | TCXO | OCXO | Rb | > |-------------------+-------+-------+-------| > | 0.01 | 1E-10 | 2E-11 | 2E-11 | > | 0.1 | 4E-11 | 4E-12 | 4E-12 | > | 1 | 3E-11 | 2E-12 | 1E-11 | > | 10 | 1E-10 | 2E-12 | 6E-12 | > | 100 | 6E-11 | 2E-12 | 1E-12 | > | 1k | 6E-12 | 3E-12 | 1E-12 | > | 10k | 6E-13 | 6E-13 | 3E-13 | > | 100k | 6E-14 | 6E-14 | 3E-14 | > | 1M | 5E-15 | 5E-15 | 3E-15 | > > (I used the typical performance plots for the phase noise and Allan > deviation data) > > The OCXO does not exhibit the the hump around 2s seen by the rubidium > oscillator option (PRS10) and therefore could be seen as better in > applications not requiring the better holdover performance by the Rb. > > It also provides a pulse output with jitter < 50 ps RMS that could be > used as a PPS signal. It can also provide a low noise 100 MHz output > signal and can act as a frequency counter and DDS with high resolution > and accuracy. And it comes with the schematics! But.... it's definitely > not cheap. Other thoughts/opinions? > > Matt > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send > an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there. >
MH
Matt Huszagh
Wed, Nov 17, 2021 3:16 PM

Bob kb8tq kb8tq@n1k.org writes:

I have a 740 as do others on the list. The big gotcha is the OCXO in the device.
It has various quirks that limit the performance of the device. The ADEV plot they
show for the OCXO is a “sometimes / maybe / might” sort of thing. One of the
“get to it eventually” projects here is to see if it can be replaced with something
that has better ADEV performance.

Good to know. Have you tried disciplining the PRS-10 in one of these
units? Based on Dana's comments, it sounds like this may not exhibit the
same ADEV susceptibility to the thermal environment.

Matt

Bob kb8tq <kb8tq@n1k.org> writes: > I have a 740 as do others on the list. The big gotcha is the OCXO in the device. > It has various quirks that limit the performance of the device. The ADEV plot they > show for the OCXO is a “sometimes / maybe / might” sort of thing. One of the > “get to it eventually” projects here is to see if it can be replaced with something > that has better ADEV performance. Good to know. Have you tried disciplining the PRS-10 in one of these units? Based on Dana's comments, it sounds like this may not exhibit the same ADEV susceptibility to the thermal environment. Matt
E
ew
Wed, Nov 17, 2021 3:21 PM

I have been looking at it for years had an email conversation with the designer and Bob. This box is not intended for time nuts and does sell very well, price has gone up over 20 percent.  It would be nice if one could use an external OCXO or Rb and using the Internet interface and have SRS download a Filter Time Constant for a fee based on data submitted. No deal.                                                                                                                                               Bert Kehren                                                                                                                                                                                                                                                                                          

In a message dated 11/17/2021 9:09:42 AM Eastern Standard Time, k8yumdoober@gmail.com writes: 
I think that before one places much faith in the ADEV data, it's important
to understand the thermal environment in which the tests were conducted,
compared to your own thermal environment.  The tests were likely to have
been performed in a well-stabilized ambient temperature, to make them
look good.  But your own results may be a lot worse due to HVAC cycling
etc.

One way to reduce such effects is to make the GPS locking loop quite fast,
but doing so is at the cost of the disciplines oscillator's being yanked
around
in phase by "GPS noise", which is really quite awful.  At the other
extreme,
(long locking loop time constant) one largely escapes GPS noise, but finds
that his system has essentially become an expensive thermometer.  So I favor
the use of a GPS-disciplined Rb for critical work (or an active H-maser if
you're
really rich).

I'm presently crutching along with a manually-disciplined surplus PRS-10,
but
am looking for a good excuse to get a factory new lockable unit.

Aha:  KB8TQ's note just came in;  we're saying about the same thing- I'm
just
wordier.

Dana  K8YUM

On Tue, Nov 16, 2021 at 11:59 PM Matt Huszagh huszaghmatt@gmail.com wrote:

Hi,

I'm curious to hear people's thoughts on the SRS FS740
(https://www.thinksrs.com/products/fs740.html) as a lab frequency
standard. The phase noise performance of the OCXO and Rb options seem
quite good (the same specs apparently apply during GPS disciplining too)

| offset |  TCXO  | OCXO/Rb  |
| (Hz)  | (dBc/Hz) | (dBc/Hz) |
|--------+----------+----------|
| 1      |  -90    |  -108  |
| 10    |  -104  |  -130  |
| 100    |  -125  |  -144  |
| 1k    |  -142  |  -143  |
| 10k    |  -147  |  -148  |
| 100k  |  -149  |  -149  |
| 1M    |  -153  |  -153  |

as does the Allan deviation (during GPS disciplining)

| time interval (s) | TCXO  | OCXO  |  Rb  |
|-------------------+-------+-------+-------|
| 0.01              | 1E-10 | 2E-11 | 2E-11 |
| 0.1              | 4E-11 | 4E-12 | 4E-12 |
| 1                | 3E-11 | 2E-12 | 1E-11 |
| 10                | 1E-10 | 2E-12 | 6E-12 |
| 100              | 6E-11 | 2E-12 | 1E-12 |
| 1k                | 6E-12 | 3E-12 | 1E-12 |
| 10k              | 6E-13 | 6E-13 | 3E-13 |
| 100k              | 6E-14 | 6E-14 | 3E-14 |
| 1M                | 5E-15 | 5E-15 | 3E-15 |

(I used the typical performance plots for the phase noise and Allan
deviation data)

The OCXO does not exhibit the the hump around 2s seen by the rubidium
oscillator option (PRS10) and therefore could be seen as better in
applications not requiring the better holdover performance by the Rb.

It also provides a pulse output with jitter < 50 ps RMS that could be
used as a PPS signal. It can also provide a low noise 100 MHz output
signal and can act as a frequency counter and DDS with high resolution
and accuracy. And it comes with the schematics! But.... it's definitely
not cheap. Other thoughts/opinions?

Matt


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

I have been looking at it for years had an email conversation with the designer and Bob. This box is not intended for time nuts and does sell very well, price has gone up over 20 percent.  It would be nice if one could use an external OCXO or Rb and using the Internet interface and have SRS download a Filter Time Constant for a fee based on data submitted. No deal.                                                                                                                                               Bert Kehren                                                                                                                                                                                                                                                                                           In a message dated 11/17/2021 9:09:42 AM Eastern Standard Time, k8yumdoober@gmail.com writes:  I think that before one places much faith in the ADEV data, it's important to understand the thermal environment in which the tests were conducted, compared to your own thermal environment.  The tests were likely to have been performed in a well-stabilized ambient temperature, to make them look good.  But your own results may be a lot worse due to HVAC cycling etc. One way to reduce such effects is to make the GPS locking loop quite fast, but doing so is at the cost of the disciplines oscillator's being yanked around in phase by "GPS noise", which is really quite awful.  At the other extreme, (long locking loop time constant) one largely escapes GPS noise, but finds that his system has essentially become an expensive thermometer.  So I favor the use of a GPS-disciplined Rb for critical work (or an active H-maser if you're really rich). I'm presently crutching along with a manually-disciplined surplus PRS-10, but am looking for a good excuse to get a factory new lockable unit. Aha:  KB8TQ's note just came in;  we're saying about the same thing- I'm just wordier. Dana  K8YUM On Tue, Nov 16, 2021 at 11:59 PM Matt Huszagh <huszaghmatt@gmail.com> wrote: > Hi, > > I'm curious to hear people's thoughts on the SRS FS740 > (https://www.thinksrs.com/products/fs740.html) as a lab frequency > standard. The phase noise performance of the OCXO and Rb options seem > quite good (the same specs apparently apply during GPS disciplining too) > > | offset |  TCXO  | OCXO/Rb  | > | (Hz)  | (dBc/Hz) | (dBc/Hz) | > |--------+----------+----------| > | 1      |  -90    |  -108  | > | 10    |  -104  |  -130  | > | 100    |  -125  |  -144  | > | 1k    |  -142  |  -143  | > | 10k    |  -147  |  -148  | > | 100k  |  -149  |  -149  | > | 1M    |  -153  |  -153  | > > as does the Allan deviation (during GPS disciplining) > > | time interval (s) | TCXO  | OCXO  |  Rb  | > |-------------------+-------+-------+-------| > | 0.01              | 1E-10 | 2E-11 | 2E-11 | > | 0.1              | 4E-11 | 4E-12 | 4E-12 | > | 1                | 3E-11 | 2E-12 | 1E-11 | > | 10                | 1E-10 | 2E-12 | 6E-12 | > | 100              | 6E-11 | 2E-12 | 1E-12 | > | 1k                | 6E-12 | 3E-12 | 1E-12 | > | 10k              | 6E-13 | 6E-13 | 3E-13 | > | 100k              | 6E-14 | 6E-14 | 3E-14 | > | 1M                | 5E-15 | 5E-15 | 3E-15 | > > (I used the typical performance plots for the phase noise and Allan > deviation data) > > The OCXO does not exhibit the the hump around 2s seen by the rubidium > oscillator option (PRS10) and therefore could be seen as better in > applications not requiring the better holdover performance by the Rb. > > It also provides a pulse output with jitter < 50 ps RMS that could be > used as a PPS signal. It can also provide a low noise 100 MHz output > signal and can act as a frequency counter and DDS with high resolution > and accuracy. And it comes with the schematics! But.... it's definitely > not cheap. Other thoughts/opinions? > > Matt > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send > an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there. > _______________________________________________ time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com To unsubscribe, go to and follow the instructions there.
BK
Bob kb8tq
Wed, Nov 17, 2021 3:34 PM

Hi

In this case it’s not thermal. I’ve tested a ton of parts in the same environment ( maybe
not 2,000 pounds but several hundred OCXO’s ).

The PRS-10 gives you that hump at 2 seconds no matter what you do. It’s part of the
basic design of the unit ….

Bob

On Nov 17, 2021, at 10:16 AM, Matt Huszagh huszaghmatt@gmail.com wrote:

Bob kb8tq kb8tq@n1k.org writes:

I have a 740 as do others on the list. The big gotcha is the OCXO in the device.
It has various quirks that limit the performance of the device. The ADEV plot they
show for the OCXO is a “sometimes / maybe / might” sort of thing. One of the
“get to it eventually” projects here is to see if it can be replaced with something
that has better ADEV performance.

Good to know. Have you tried disciplining the PRS-10 in one of these
units? Based on Dana's comments, it sounds like this may not exhibit the
same ADEV susceptibility to the thermal environment.

Matt

Hi In this case it’s not thermal. I’ve tested a ton of parts in the same environment ( maybe not 2,000 pounds but several hundred OCXO’s ). The PRS-10 gives you that hump at 2 seconds no matter what you do. It’s part of the basic design of the unit …. Bob > On Nov 17, 2021, at 10:16 AM, Matt Huszagh <huszaghmatt@gmail.com> wrote: > > Bob kb8tq <kb8tq@n1k.org> writes: > >> I have a 740 as do others on the list. The big gotcha is the OCXO in the device. >> It has various quirks that limit the performance of the device. The ADEV plot they >> show for the OCXO is a “sometimes / maybe / might” sort of thing. One of the >> “get to it eventually” projects here is to see if it can be replaced with something >> that has better ADEV performance. > > Good to know. Have you tried disciplining the PRS-10 in one of these > units? Based on Dana's comments, it sounds like this may not exhibit the > same ADEV susceptibility to the thermal environment. > > Matt
DW
Dana Whitlow
Wed, Nov 17, 2021 9:24 PM

But if you can stand the ADEV hump around 2 sec, the PRS-10 is delightful.

BTW, the ordinary factory new PRS-10 does provide for direct locking to 1
PPS,
but many of the available surplus units had that feature removed.  Tne new
PRS-10's sold for years for $1595, but I see that it has gone up to $1695
since
I had last checked.

Dana

On Wed, Nov 17, 2021 at 9:52 AM Bob kb8tq kb8tq@n1k.org wrote:

Hi

In this case it’s not thermal. I’ve tested a ton of parts in the same
environment ( maybe
not 2,000 pounds but several hundred OCXO’s ).

The PRS-10 gives you that hump at 2 seconds no matter what you do. It’s
part of the
basic design of the unit ….

Bob

On Nov 17, 2021, at 10:16 AM, Matt Huszagh huszaghmatt@gmail.com

wrote:

Bob kb8tq kb8tq@n1k.org writes:

I have a 740 as do others on the list. The big gotcha is the OCXO in

the device.

It has various quirks that limit the performance of the device. The

ADEV plot they

show for the OCXO is a “sometimes / maybe / might” sort of thing. One

of the

“get to it eventually” projects here is to see if it can be replaced

with something

that has better ADEV performance.

Good to know. Have you tried disciplining the PRS-10 in one of these
units? Based on Dana's comments, it sounds like this may not exhibit the
same ADEV susceptibility to the thermal environment.

Matt


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

But if you can stand the ADEV hump around 2 sec, the PRS-10 is delightful. BTW, the ordinary factory new PRS-10 does provide for direct locking to 1 PPS, but many of the available surplus units had that feature removed. Tne new PRS-10's sold for years for $1595, but I see that it has gone up to $1695 since I had last checked. Dana On Wed, Nov 17, 2021 at 9:52 AM Bob kb8tq <kb8tq@n1k.org> wrote: > Hi > > In this case it’s not thermal. I’ve tested a ton of parts in the same > environment ( maybe > not 2,000 pounds but several hundred OCXO’s ). > > The PRS-10 gives you that hump at 2 seconds no matter what you do. It’s > part of the > basic design of the unit …. > > Bob > > > On Nov 17, 2021, at 10:16 AM, Matt Huszagh <huszaghmatt@gmail.com> > wrote: > > > > Bob kb8tq <kb8tq@n1k.org> writes: > > > >> I have a 740 as do others on the list. The big gotcha is the OCXO in > the device. > >> It has various quirks that limit the performance of the device. The > ADEV plot they > >> show for the OCXO is a “sometimes / maybe / might” sort of thing. One > of the > >> “get to it eventually” projects here is to see if it can be replaced > with something > >> that has better ADEV performance. > > > > Good to know. Have you tried disciplining the PRS-10 in one of these > > units? Based on Dana's comments, it sounds like this may not exhibit the > > same ADEV susceptibility to the thermal environment. > > > > Matt > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send > an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there.
MH
Matt Huszagh
Wed, Nov 17, 2021 9:45 PM

Dana Whitlow k8yumdoober@gmail.com writes:

But if you can stand the ADEV hump around 2 sec, the PRS-10 is delightful.

BTW, the ordinary factory new PRS-10 does provide for direct locking to 1
PPS,
but many of the available surplus units had that feature removed.  Tne new
PRS-10's sold for years for $1595, but I see that it has gone up to $1695
since
I had last checked.

Have you had luck steering the PRS10 with GPS? One individual mentioned
to me that despite repeated efforts, he was unable to discipline the
PRS10 without significantly degrading the short-term stability and phase
noise.

However, if the performance as reported by SRS isn't just for some
impossible, ideal environment than one solution would just be to
replicate the circuitry they use for the loop filter and divider to
generate the 10 MHz PPS signal. Or, to just purchase the FS740.

Matt

Dana Whitlow <k8yumdoober@gmail.com> writes: > But if you can stand the ADEV hump around 2 sec, the PRS-10 is delightful. > > BTW, the ordinary factory new PRS-10 does provide for direct locking to 1 > PPS, > but many of the available surplus units had that feature removed. Tne new > PRS-10's sold for years for $1595, but I see that it has gone up to $1695 > since > I had last checked. Have you had luck steering the PRS10 with GPS? One individual mentioned to me that despite repeated efforts, he was unable to discipline the PRS10 without significantly degrading the short-term stability and phase noise. However, if the performance as reported by SRS isn't just for some impossible, ideal environment than one solution would just be to replicate the circuitry they use for the loop filter and divider to generate the 10 MHz PPS signal. Or, to just purchase the FS740. Matt
BC
Brooke Clarke
Wed, Nov 17, 2021 10:15 PM

Hi Dana:

My PRS10 is a couple decades old and I use it in the 1PPS locking mode.  But that may not have been the default.  Maybe
the surplus units just need the correct jumper change.
https://prc68.com/I/PRS10.shtml

--
Have Fun,

Brooke Clarke
https://www.PRC68.com
axioms:

  1. The extent to which you can fix or improve something will be limited by how well you understand how it works.
  2. Everybody, with no exceptions, holds false beliefs.

-------- Original Message --------

But if you can stand the ADEV hump around 2 sec, the PRS-10 is delightful.

BTW, the ordinary factory new PRS-10 does provide for direct locking to 1
PPS,
but many of the available surplus units had that feature removed.  Tne new
PRS-10's sold for years for $1595, but I see that it has gone up to $1695
since
I had last checked.

Dana

On Wed, Nov 17, 2021 at 9:52 AM Bob kb8tq kb8tq@n1k.org wrote:

Hi

In this case it’s not thermal. I’ve tested a ton of parts in the same
environment ( maybe
not 2,000 pounds but several hundred OCXO’s ).

The PRS-10 gives you that hump at 2 seconds no matter what you do. It’s
part of the
basic design of the unit ….

Bob

On Nov 17, 2021, at 10:16 AM, Matt Huszagh huszaghmatt@gmail.com

wrote:

Bob kb8tq kb8tq@n1k.org writes:

I have a 740 as do others on the list. The big gotcha is the OCXO in

the device.

It has various quirks that limit the performance of the device. The

ADEV plot they

show for the OCXO is a “sometimes / maybe / might” sort of thing. One

of the

“get to it eventually” projects here is to see if it can be replaced

with something

that has better ADEV performance.

Good to know. Have you tried disciplining the PRS-10 in one of these
units? Based on Dana's comments, it sounds like this may not exhibit the
same ADEV susceptibility to the thermal environment.

Matt


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

Hi Dana: My PRS10 is a couple decades old and I use it in the 1PPS locking mode.  But that may not have been the default.  Maybe the surplus units just need the correct jumper change. https://prc68.com/I/PRS10.shtml -- Have Fun, Brooke Clarke https://www.PRC68.com axioms: 1. The extent to which you can fix or improve something will be limited by how well you understand how it works. 2. Everybody, with no exceptions, holds false beliefs. -------- Original Message -------- > But if you can stand the ADEV hump around 2 sec, the PRS-10 is delightful. > > BTW, the ordinary factory new PRS-10 does provide for direct locking to 1 > PPS, > but many of the available surplus units had that feature removed. Tne new > PRS-10's sold for years for $1595, but I see that it has gone up to $1695 > since > I had last checked. > > Dana > > > On Wed, Nov 17, 2021 at 9:52 AM Bob kb8tq <kb8tq@n1k.org> wrote: > >> Hi >> >> In this case it’s not thermal. I’ve tested a ton of parts in the same >> environment ( maybe >> not 2,000 pounds but several hundred OCXO’s ). >> >> The PRS-10 gives you that hump at 2 seconds no matter what you do. It’s >> part of the >> basic design of the unit …. >> >> Bob >> >>> On Nov 17, 2021, at 10:16 AM, Matt Huszagh <huszaghmatt@gmail.com> >> wrote: >>> Bob kb8tq <kb8tq@n1k.org> writes: >>> >>>> I have a 740 as do others on the list. The big gotcha is the OCXO in >> the device. >>>> It has various quirks that limit the performance of the device. The >> ADEV plot they >>>> show for the OCXO is a “sometimes / maybe / might” sort of thing. One >> of the >>>> “get to it eventually” projects here is to see if it can be replaced >> with something >>>> that has better ADEV performance. >>> Good to know. Have you tried disciplining the PRS-10 in one of these >>> units? Based on Dana's comments, it sounds like this may not exhibit the >>> same ADEV susceptibility to the thermal environment. >>> >>> Matt >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send >> an email to time-nuts-leave@lists.febo.com >> To unsubscribe, go to and follow the instructions there. > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there.
PK
Poul-Henning Kamp
Thu, Nov 18, 2021 7:08 AM

Dana Whitlow writes:

But if you can stand the ADEV hump around 2 sec, the PRS-10 is delightful.

The PRS-10 switches the polarity of the C-field solenoid at 5Hz to
cancel out varying external magnetic fields.

If that is not a concern for you, for instance because you use it
in a stationary application, it can be disabled with the "MS0" command.

It's all in the manual.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

-------- Dana Whitlow writes: > But if you can stand the ADEV hump around 2 sec, the PRS-10 is delightful. The PRS-10 switches the polarity of the C-field solenoid at 5Hz to cancel out varying external magnetic fields. If that is not a concern for you, for instance because you use it in a stationary application, it can be disabled with the "MS0" command. It's all in the manual. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
MH
Matt Huszagh
Thu, Nov 18, 2021 7:32 AM

"Poul-Henning Kamp" phk@phk.freebsd.dk writes:

The PRS-10 switches the polarity of the C-field solenoid at 5Hz to
cancel out varying external magnetic fields.

If that is not a concern for you, for instance because you use it
in a stationary application, it can be disabled with the "MS0" command.

It's all in the manual.

How does the 5Hz switching relate to the 2s hump in the ADEV plot?

Matt

"Poul-Henning Kamp" <phk@phk.freebsd.dk> writes: > The PRS-10 switches the polarity of the C-field solenoid at 5Hz to > cancel out varying external magnetic fields. > > If that is not a concern for you, for instance because you use it > in a stationary application, it can be disabled with the "MS0" command. > > It's all in the manual. How does the 5Hz switching relate to the 2s hump in the ADEV plot? Matt
PK
Poul-Henning Kamp
Thu, Nov 18, 2021 8:41 AM

Matt Huszagh writes:

"Poul-Henning Kamp" phk@phk.freebsd.dk writes:

The PRS-10 switches the polarity of the C-field solenoid at 5Hz to
cancel out varying external magnetic fields.

If that is not a concern for you, for instance because you use it
in a stationary application, it can be disabled with the "MS0" command.

It's all in the manual.

How does the 5Hz switching relate to the 2s hump in the ADEV plot?

The manual says:

"[...]the current in the coil is switched at a 5 Hz rate."

You can either read that as:

"There are five positive and five negative periods every second"

or
"The sign changes five times per second"

It is not entirely obvious which reading is the correct one.

When I experimented with it ages ago, I concluded the latter fit
my data best, but that was a pretty early firmware version, with
quite a number of variances from the manual.

If your ambient magnetic field is stable, and it should be for
time-nuts purposes, modulating the hyperfine transition is a bad
idea, no matter the frequency.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

-------- Matt Huszagh writes: > "Poul-Henning Kamp" <phk@phk.freebsd.dk> writes: > > > The PRS-10 switches the polarity of the C-field solenoid at 5Hz to > > cancel out varying external magnetic fields. > > > > If that is not a concern for you, for instance because you use it > > in a stationary application, it can be disabled with the "MS0" command. > > > > It's all in the manual. > > How does the 5Hz switching relate to the 2s hump in the ADEV plot? The manual says: "[...]the current in the coil is switched at a 5 Hz rate." You can either read that as: "There are five positive and five negative periods every second" or "The sign changes five times per second" It is not entirely obvious which reading is the correct one. When I experimented with it ages ago, I concluded the latter fit my data best, but that was a pretty early firmware version, with quite a number of variances from the manual. If your ambient magnetic field is stable, and it should be for time-nuts purposes, modulating the hyperfine transition is a bad idea, no matter the frequency. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
DW
Dana Whitlow
Thu, Nov 18, 2021 12:50 PM

Poul,

You did not really answer the basic question: "Does the 2.5 Hz
(or 5 Hz if applicable) C-field reversal cause the 2-sec bump in
the ADEV plot for  the PRS-10?".

BTW, some radio hams are involved with microwave and even
MMW communications between mobile stations, and I'd bet they
benefit quite a bit from any scheme that reduces magnetic field
sensitivity of their frequency references.  Therefore, the C-field
reversal thing cannot be all bad.

Dana    K8YUM

Dana

On Thu, Nov 18, 2021 at 2:41 AM Poul-Henning Kamp phk@phk.freebsd.dk
wrote:


Matt Huszagh writes:

"Poul-Henning Kamp" phk@phk.freebsd.dk writes:

The PRS-10 switches the polarity of the C-field solenoid at 5Hz to
cancel out varying external magnetic fields.

If that is not a concern for you, for instance because you use it
in a stationary application, it can be disabled with the "MS0" command.

It's all in the manual.

How does the 5Hz switching relate to the 2s hump in the ADEV plot?

The manual says:

     "[...]the current in the coil is switched at a 5 Hz rate."

You can either read that as:

     "There are five positive and five negative periods every second"

or
"The sign changes five times per second"

It is not entirely obvious which reading is the correct one.

When I experimented with it ages ago, I concluded the latter fit
my data best, but that was a pretty early firmware version, with
quite a number of variances from the manual.

If your ambient magnetic field is stable, and it should be for
time-nuts purposes, modulating the hyperfine transition is a bad
idea, no matter the frequency.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

Poul, You did not really answer the basic question: "Does the 2.5 Hz (or 5 Hz if applicable) C-field reversal cause the 2-sec bump in the ADEV plot for the PRS-10?". BTW, some radio hams are involved with microwave and even MMW communications between mobile stations, and I'd bet they benefit quite a bit from any scheme that reduces magnetic field sensitivity of their frequency references. Therefore, the C-field reversal thing cannot be *all* bad. Dana K8YUM Dana On Thu, Nov 18, 2021 at 2:41 AM Poul-Henning Kamp <phk@phk.freebsd.dk> wrote: > -------- > Matt Huszagh writes: > > "Poul-Henning Kamp" <phk@phk.freebsd.dk> writes: > > > > > The PRS-10 switches the polarity of the C-field solenoid at 5Hz to > > > cancel out varying external magnetic fields. > > > > > > If that is not a concern for you, for instance because you use it > > > in a stationary application, it can be disabled with the "MS0" command. > > > > > > It's all in the manual. > > > > How does the 5Hz switching relate to the 2s hump in the ADEV plot? > > The manual says: > > "[...]the current in the coil is switched at a 5 Hz rate." > > You can either read that as: > > "There are five positive and five negative periods every second" > > or > "The sign changes five times per second" > > It is not entirely obvious which reading is the correct one. > > When I experimented with it ages ago, I concluded the latter fit > my data best, but that was a pretty early firmware version, with > quite a number of variances from the manual. > > If your ambient magnetic field is stable, and it should be for > time-nuts purposes, modulating the hyperfine transition is a bad > idea, no matter the frequency. > > -- > Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 > phk@FreeBSD.ORG | TCP/IP since RFC 956 > FreeBSD committer | BSD since 4.3-tahoe > Never attribute to malice what can adequately be explained by incompetence. >
PK
Poul-Henning Kamp
Thu, Nov 18, 2021 1:01 PM

Dana Whitlow writes:

You did not really answer the basic question: "Does the 2.5 Hz
(or 5 Hz if applicable) C-field reversal cause the 2-sec bump in
the ADEV plot for  the PRS-10?".

I cannot answer that question as stated.

My PRS10 died years ago, all I can say is that when I commissioned
it, I disabled the MS because it improved things.

BTW, some radio hams are involved with microwave and even
MMW communications between mobile stations, [...]

In any mobile application, or other application where the magnetic
field is not nearly contant (for instance behind a steel door in a
outdoor base-station cabinet, I would absolutely keep MS on.

But for a time-nut installation, where the location is fixed,
temperature controlled and, I would assume, due consideration
is given to also keep the magnetic field from slamming around,
I suggest MS be disabled.

But it's a really trivial experiment to perform, so do it, measure
and decide what is right for you.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

-------- Dana Whitlow writes: > You did not really answer the basic question: "Does the 2.5 Hz > (or 5 Hz if applicable) C-field reversal cause the 2-sec bump in > the ADEV plot for the PRS-10?". I cannot answer that question as stated. My PRS10 died years ago, all I can say is that when I commissioned it, I disabled the MS because it improved things. > BTW, some radio hams are involved with microwave and even > MMW communications between mobile stations, [...] In any mobile application, or other application where the magnetic field is not nearly contant (for instance behind a steel door in a outdoor base-station cabinet, I would absolutely keep MS on. But for a time-nut installation, where the location is fixed, temperature controlled and, I would assume, due consideration is given to also keep the magnetic field from slamming around, I suggest MS be disabled. But it's a really trivial experiment to perform, so do it, measure and decide what is right for you. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
BK
Bob kb8tq
Thu, Nov 18, 2021 2:13 PM

Hi

The 5 Hz (or whatever) switching and the hump definitely both
originate in the same “feature” of the design. Does disabling the
switching nuke the hump? If one has a PRS-10 I’d say it’s well
worth trying.

The “filter” in the PRS-10 is not really set up for a GPS sort of
signal. It’s designed to handle a nice clean lab generated 1 pps.
It’s there to lock the device up as part of a calibration routine.
The 1 pps sync setups on other Rb’s are designed with the same
target in mind.

Bob

On Nov 18, 2021, at 7:50 AM, Dana Whitlow k8yumdoober@gmail.com wrote:

Poul,

You did not really answer the basic question: "Does the 2.5 Hz
(or 5 Hz if applicable) C-field reversal cause the 2-sec bump in
the ADEV plot for  the PRS-10?".

BTW, some radio hams are involved with microwave and even
MMW communications between mobile stations, and I'd bet they
benefit quite a bit from any scheme that reduces magnetic field
sensitivity of their frequency references.  Therefore, the C-field
reversal thing cannot be all bad.

Dana    K8YUM

Dana

On Thu, Nov 18, 2021 at 2:41 AM Poul-Henning Kamp phk@phk.freebsd.dk
wrote:


Matt Huszagh writes:

"Poul-Henning Kamp" phk@phk.freebsd.dk writes:

The PRS-10 switches the polarity of the C-field solenoid at 5Hz to
cancel out varying external magnetic fields.

If that is not a concern for you, for instance because you use it
in a stationary application, it can be disabled with the "MS0" command.

It's all in the manual.

How does the 5Hz switching relate to the 2s hump in the ADEV plot?

The manual says:

    "[...]the current in the coil is switched at a 5 Hz rate."

You can either read that as:

    "There are five positive and five negative periods every second"

or
"The sign changes five times per second"

It is not entirely obvious which reading is the correct one.

When I experimented with it ages ago, I concluded the latter fit
my data best, but that was a pretty early firmware version, with
quite a number of variances from the manual.

If your ambient magnetic field is stable, and it should be for
time-nuts purposes, modulating the hyperfine transition is a bad
idea, no matter the frequency.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

Hi The 5 Hz (or whatever) switching and the hump definitely both originate in the same “feature” of the design. Does disabling the switching nuke the hump? If one has a PRS-10 I’d say it’s well worth trying. The “filter” in the PRS-10 is not really set up for a GPS sort of signal. It’s designed to handle a nice clean lab generated 1 pps. It’s there to lock the device up as part of a calibration routine. The 1 pps sync setups on other Rb’s are designed with the same target in mind. Bob > On Nov 18, 2021, at 7:50 AM, Dana Whitlow <k8yumdoober@gmail.com> wrote: > > Poul, > > You did not really answer the basic question: "Does the 2.5 Hz > (or 5 Hz if applicable) C-field reversal cause the 2-sec bump in > the ADEV plot for the PRS-10?". > > BTW, some radio hams are involved with microwave and even > MMW communications between mobile stations, and I'd bet they > benefit quite a bit from any scheme that reduces magnetic field > sensitivity of their frequency references. Therefore, the C-field > reversal thing cannot be *all* bad. > > Dana K8YUM > > > Dana > > > On Thu, Nov 18, 2021 at 2:41 AM Poul-Henning Kamp <phk@phk.freebsd.dk> > wrote: > >> -------- >> Matt Huszagh writes: >>> "Poul-Henning Kamp" <phk@phk.freebsd.dk> writes: >>> >>>> The PRS-10 switches the polarity of the C-field solenoid at 5Hz to >>>> cancel out varying external magnetic fields. >>>> >>>> If that is not a concern for you, for instance because you use it >>>> in a stationary application, it can be disabled with the "MS0" command. >>>> >>>> It's all in the manual. >>> >>> How does the 5Hz switching relate to the 2s hump in the ADEV plot? >> >> The manual says: >> >> "[...]the current in the coil is switched at a 5 Hz rate." >> >> You can either read that as: >> >> "There are five positive and five negative periods every second" >> >> or >> "The sign changes five times per second" >> >> It is not entirely obvious which reading is the correct one. >> >> When I experimented with it ages ago, I concluded the latter fit >> my data best, but that was a pretty early firmware version, with >> quite a number of variances from the manual. >> >> If your ambient magnetic field is stable, and it should be for >> time-nuts purposes, modulating the hyperfine transition is a bad >> idea, no matter the frequency. >> >> -- >> Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 >> phk@FreeBSD.ORG | TCP/IP since RFC 956 >> FreeBSD committer | BSD since 4.3-tahoe >> Never attribute to malice what can adequately be explained by incompetence. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there.
PK
Poul-Henning Kamp
Thu, Nov 18, 2021 2:41 PM

Bob kb8tq writes:

The 'filter' in the PRS-10 is not really set up for a GPS sort of signal.

It is particularly bad at handling GPS's because of the "hanging bridge"
phenomena, and the better the GPS, the worse the result...

I tried injecting the "negative sawtooth" via the serial port to my
PRS10 but firmware features/bugs prevented that.

The offset could only be changed permanently in the saved configuration,
you could not change the running value on a second to second basis.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

-------- Bob kb8tq writes: > The 'filter' in the PRS-10 is not really set up for a GPS sort of signal. It is particularly bad at handling GPS's because of the "hanging bridge" phenomena, and the better the GPS, the worse the result... I tried injecting the "negative sawtooth" via the serial port to my PRS10 but firmware features/bugs prevented that. The offset could only be changed permanently in the saved configuration, you could not change the running value on a second to second basis. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
DW
Dana Whitlow
Thu, Nov 18, 2021 7:02 PM

Bob, I seem to remember reading in the manual on the PRS-10 where they
expounded
at length about how they took great pains to make the unit work well with
typical noisy
GPS PPS input.

I'm using a CNS Clock II as my primary GPSDO, checking phase slippage with
respect
to my PRS-10 to make decisions about when (or when not) to manually tweak
the PRS-10.
The CNS seems awfully noisy to me, even though the mfr says that it has
strong hanging
bridge correction.  I run the two 10 MHz signals through a quadrature
demodulator whose
I & Q outputs go into a 2-channel DSO running in extremely low "roll
mode".  I take a glance
at it several times per day, with each glance showing me the most recent
~4-hour history.
When I do tweak, I strive to set the PRS-10 about 1E-11 low in frequency,
which yields a
couple of months of hands-off operation before it drifts to 1E-11 on the
high side.  For
now, that has to suffice.  But I'd rather not have to pay so much attention
to it, which is
why I dream about buying a new PRS-10 with PPS locking capability.

Dana

On Thu, Nov 18, 2021 at 8:47 AM Poul-Henning Kamp phk@phk.freebsd.dk
wrote:


Bob kb8tq writes:

The 'filter' in the PRS-10 is not really set up for a GPS sort of signal.

It is particularly bad at handling GPS's because of the "hanging bridge"
phenomena, and the better the GPS, the worse the result...

I tried injecting the "negative sawtooth" via the serial port to my
PRS10 but firmware features/bugs prevented that.

The offset could only be changed permanently in the saved configuration,
you could not change the running value on a second to second basis.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

Bob, I seem to remember reading in the manual on the PRS-10 where they expounded at length about how they took great pains to make the unit work well with typical noisy GPS PPS input. I'm using a CNS Clock II as my primary GPSDO, checking phase slippage with respect to my PRS-10 to make decisions about when (or when not) to manually tweak the PRS-10. The CNS seems awfully noisy to me, even though the mfr says that it has strong hanging bridge correction. I run the two 10 MHz signals through a quadrature demodulator whose I & Q outputs go into a 2-channel DSO running in extremely low "roll mode". I take a glance at it several times per day, with each glance showing me the most recent ~4-hour history. When I do tweak, I strive to set the PRS-10 about 1E-11 low in frequency, which yields a couple of months of hands-off operation before it drifts to 1E-11 on the high side. For now, that has to suffice. But I'd rather not have to pay so much attention to it, which is why I dream about buying a new PRS-10 with PPS locking capability. Dana On Thu, Nov 18, 2021 at 8:47 AM Poul-Henning Kamp <phk@phk.freebsd.dk> wrote: > -------- > Bob kb8tq writes: > > > The 'filter' in the PRS-10 is not really set up for a GPS sort of signal. > > It is particularly bad at handling GPS's because of the "hanging bridge" > phenomena, and the better the GPS, the worse the result... > > I tried injecting the "negative sawtooth" via the serial port to my > PRS10 but firmware features/bugs prevented that. > > The offset could only be changed permanently in the saved configuration, > you could not change the running value on a second to second basis. > > -- > Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 > phk@FreeBSD.ORG | TCP/IP since RFC 956 > FreeBSD committer | BSD since 4.3-tahoe > Never attribute to malice what can adequately be explained by incompetence. > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send > an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there. >
BK
Bob kb8tq
Thu, Nov 18, 2021 7:12 PM

Hi

This is one of those classic examples of what marketing writes and what
(once the beer is poured) engineering actually knows. If you dive into the
grubby details of the filtering. Been there / done that with a number of
these guys.

It’s very much not set up for GPS. The hanging bridge stuff PHK mentions
in a previous post is one example. The need to adapt the filter to the monitored
results is another weak area. Just how and when this or that pops up varies,
but it will pop up.

Bob

On Nov 18, 2021, at 2:02 PM, Dana Whitlow k8yumdoober@gmail.com wrote:

Bob, I seem to remember reading in the manual on the PRS-10 where they
expounded
at length about how they took great pains to make the unit work well with
typical noisy
GPS PPS input.

I'm using a CNS Clock II as my primary GPSDO, checking phase slippage with
respect
to my PRS-10 to make decisions about when (or when not) to manually tweak
the PRS-10.
The CNS seems awfully noisy to me, even though the mfr says that it has
strong hanging
bridge correction.  I run the two 10 MHz signals through a quadrature
demodulator whose
I & Q outputs go into a 2-channel DSO running in extremely low "roll
mode".  I take a glance
at it several times per day, with each glance showing me the most recent
~4-hour history.
When I do tweak, I strive to set the PRS-10 about 1E-11 low in frequency,
which yields a
couple of months of hands-off operation before it drifts to 1E-11 on the
high side.  For
now, that has to suffice.  But I'd rather not have to pay so much attention
to it, which is
why I dream about buying a new PRS-10 with PPS locking capability.

Dana

On Thu, Nov 18, 2021 at 8:47 AM Poul-Henning Kamp phk@phk.freebsd.dk
wrote:


Bob kb8tq writes:

The 'filter' in the PRS-10 is not really set up for a GPS sort of signal.

It is particularly bad at handling GPS's because of the "hanging bridge"
phenomena, and the better the GPS, the worse the result...

I tried injecting the "negative sawtooth" via the serial port to my
PRS10 but firmware features/bugs prevented that.

The offset could only be changed permanently in the saved configuration,
you could not change the running value on a second to second basis.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

Hi This is one of those classic examples of what marketing writes and what (once the beer is poured) engineering actually knows. If you dive into the grubby details of the filtering. Been there / done that with a number of these guys. It’s very much *not* set up for GPS. The hanging bridge stuff PHK mentions in a previous post is one example. The need to adapt the filter to the monitored results is another weak area. Just how and when this or that pops up varies, but it will pop up. Bob > On Nov 18, 2021, at 2:02 PM, Dana Whitlow <k8yumdoober@gmail.com> wrote: > > Bob, I seem to remember reading in the manual on the PRS-10 where they > expounded > at length about how they took great pains to make the unit work well with > typical noisy > GPS PPS input. > > I'm using a CNS Clock II as my primary GPSDO, checking phase slippage with > respect > to my PRS-10 to make decisions about when (or when not) to manually tweak > the PRS-10. > The CNS seems awfully noisy to me, even though the mfr says that it has > strong hanging > bridge correction. I run the two 10 MHz signals through a quadrature > demodulator whose > I & Q outputs go into a 2-channel DSO running in extremely low "roll > mode". I take a glance > at it several times per day, with each glance showing me the most recent > ~4-hour history. > When I do tweak, I strive to set the PRS-10 about 1E-11 low in frequency, > which yields a > couple of months of hands-off operation before it drifts to 1E-11 on the > high side. For > now, that has to suffice. But I'd rather not have to pay so much attention > to it, which is > why I dream about buying a new PRS-10 with PPS locking capability. > > Dana > > > On Thu, Nov 18, 2021 at 8:47 AM Poul-Henning Kamp <phk@phk.freebsd.dk> > wrote: > >> -------- >> Bob kb8tq writes: >> >>> The 'filter' in the PRS-10 is not really set up for a GPS sort of signal. >> >> It is particularly bad at handling GPS's because of the "hanging bridge" >> phenomena, and the better the GPS, the worse the result... >> >> I tried injecting the "negative sawtooth" via the serial port to my >> PRS10 but firmware features/bugs prevented that. >> >> The offset could only be changed permanently in the saved configuration, >> you could not change the running value on a second to second basis. >> >> -- >> Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 >> phk@FreeBSD.ORG | TCP/IP since RFC 956 >> FreeBSD committer | BSD since 4.3-tahoe >> Never attribute to malice what can adequately be explained by incompetence. >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send >> an email to time-nuts-leave@lists.febo.com >> To unsubscribe, go to and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there.