New to the list, I got a used LPRO 101 and developed a GPS-based
counter to check accuracy of calibration. I found it is almost on the
spot, freq=9.99999998 MHz.
However, I can not adjust the frequency neither by adjusting the
external C-field potentiometer nor by applying a voltage at the EXT.
C-FIELD VOLTAGE pin.
Am I missing something, or do I have a faulty unit? In this case are
there schematics somewhere, or someone have experience repairing the
unit?
Geraldo
geraldo@decampos.net
Geraldo Lino de Campos said the following on 10/09/2007 05:42 PM:
); SAEximRunCond expanded to false
Errors-To: time-nuts-bounces+jra=fluffles.febo.com@febo.com RETRY
New to the list, I got a used LPRO 101 and developed a GPS-based
counter to check accuracy of calibration. I found it is almost on the
spot, freq=9.99999998 MHz.
However, I can not adjust the frequency neither by adjusting the
external C-field potentiometer nor by applying a voltage at the EXT.
C-FIELD VOLTAGE pin.
Am I missing something, or do I have a faulty unit? In this case are
there schematics somewhere, or someone have experience repairing the
unit?
Hi Geraldo. Welcome to the list!
I would suspect either the counter or the LPRO to be faulty. Can you
check either against another device?
My reason for saying so is that the 9 digit number you gave us provides
resolution only to parts in 10e-9, and the last digit of most counters
is ambiguous by nature.
A typical Rb may have a total adjustment range of only a couple of parts
in 10e-9, so your counter resolution plus last digit ambiguity pretty
much hides any change that you might see. If you can extend your gate
time by 10, that would add another digit of resolution and with that,
you should be able to see a change over the Rb tuning range. (But note
that if you are clocking the counter directly from GPS, any gate time
shorter than thousands of seconds will have jitter measured in nanoseconds.)
I would not have been surprised if your counter read either 1 count
below 10 MHz, or 1 count above -- that would be expected due to the last
digit ambiguity. But given the adjustment range, reading 2 counts low
with no change over the adjustment range does seem to indicate a problem
as that is (probably) out of the adjustment range. However, unless we
can verify that the counter is in proper order, we can't tell whether
the problem is the counter or the Rb.
Hope this helps.
John
Easier way to read those last digits is to use a synthesized 20GHz or higher
synthesizer and a counter that can read this to 1Hz. Then lock one to one source
and the other to the other source. Parts in 1E-10 now readable in a second.
73,
Jeffrey Pawlan WA6KBL
Not too long ago there was a short thread on ageing of modern crystal
oscillators. There was also a comment that most of the aging process now is
due to what is known as microcracks. I just cannot seem to find that post
now. It may have been from Rick. I wonder if the poster of that comment can
provide a good reference to that topic. I would also like to see the
original post again and would appreciate it if someone, without going
through any trouble, could forward it to me. Thanks - Mike
Mike B. Feher, N4FS
89 Arnold Blvd.
Howell, NJ, 07731
732-886-5960
It was Rick who wrote about crystal aging. It is not predictable whether the
crystal will go higher or go lower in frequency and there is no guarantee that
it will age at a linear rate. The worst-case is when a crystal erratically
jumps.
I had attended a conference of the Frequency Control Symposium of the IEEE in
which the causes were discussed. There was disagreement among the experts. Some
thought it was owing to the physical stress or surface imperfections caused by
grinding and lapping. Others thought it may be partially a result of the
etching. Others thought it was caused by microscopic impurities in the quartz.
Others talked about the effects of plating. Rick wrote about the
microcrystalline structural defects. Perhaps all of these contribute and it is
unpredictable how they will add together.
73,
Jeffrey Pawlan WA6KBL
I was the person who posted the original comments. My comments
were based on discussions with notable crystal guru, Jack Kusters,
who was merely an inventor of the SC cut. He didn't have much
confidence in the various FCS papers Jeffrey was referring to.
Basically, he was aware of all those effects and if there were
any consistency to the proposed treatments, the HP crystal fab
would have been doing them. For instance, the idea that, if
only you polish the crystal well enough, it won't age has been
debunked. Some people currently making high quality crystals
learned much of what they are doing from Jack in the first place.
If any of them go beyond what Jack taught them, they might
publish at FCS. Or they might not and keep it as a trade secret.
Rick Karlquist
(these are my personal opinions, and do not represent the official
position of Jack or HP.).
effrey Pawlan wrote:
); SAEximRunCond expanded to false
Errors-To: time-nuts-bounces+richard=karlquist.com@febo.com RETRY
It was Rick who wrote about crystal aging. It is not predictable whether
the
crystal will go higher or go lower in frequency and there is no guarantee
that
it will age at a linear rate. The worst-case is when a crystal erratically
jumps.
I had attended a conference of the Frequency Control Symposium of the IEEE
in
which the causes were discussed. There was disagreement among the experts.
Some
thought it was owing to the physical stress or surface imperfections
caused by
grinding and lapping. Others thought it may be partially a result of the
etching. Others thought it was caused by microscopic impurities in the
quartz.
Others talked about the effects of plating. Rick wrote about the
microcrystalline structural defects. Perhaps all of these contribute and
it is
unpredictable how they will add together.
73,
Jeffrey Pawlan WA6KBL
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Jeffrey Pawlan wrote:
It was Rick who wrote about crystal aging. It is not predictable whether the
crystal will go higher or go lower in frequency and there is no guarantee that
it will age at a linear rate. The worst-case is when a crystal erratically
jumps.
Indeed, the observed crystal aging is a sum of a larger number of
physical processes.
O.K. you now can say: why not make aging prediction by mathematical
fitting of a sum of logarithmic terms through the observed frequency
data over time?
Basically this would/could improve the degree of fitting, correct. BUT:
There is a big leveraging effect, if you make a mathematical prediction
over a long period of time from data taken over a relatively small time
period. In other words: If you just vary one or a few data points by a
very small amount (i.e. to eliminate jumps), the effect on the
extrapolated curve is enormous. Just try it using EXCEL's solver ...
I had published an example for that 10 years ago - see a copy of my
paper " Correlation of predicted and real aging behaviour" on our
website www.axtal.com.
In that paper I used data from about 9 months of aging, and compared the
predictions made from different time periods with the real aging.
From my standpoint the MIL-PRF55310 method of prediction is just a
standardized method to define how the aging numbers in a spec are
"verified" in a reasonable time of 4 weeks, It does not say: "This is
how the crystal will age over decades"
There is much more to say, but I will stop here, hoping that this
contribution gave some time nuts a better understanding - and may kill
their firm believe into aging predictions ;-)
Best regards
Bernd Neubig
DK1AG