Phase measurement of my GPSDO

BQ
Bob Q
Fri, Mar 20, 2020 4:16 PM

I have seen differences between both UCT and Oscilloquartz 8663 ocxo’s. The attached plot shows an example. Both boxes use Ublox LEA-6T receiver, surveyed in, AD5680 DAC 18 bit DAC, same level shift circuit and same control circuit. The reference is an LPRO-101. The Oscilloquartz ocxo was purchased used. Both UCT ocxo’s (only the better one is shown) were purchased new and have 100’s of operating hours. I have also seen differences with constant EFC control voltage. The differences limit what  performance you can achieve.

I have seen differences between both UCT and Oscilloquartz 8663 ocxo’s. The attached plot shows an example. Both boxes use Ublox LEA-6T receiver, surveyed in, AD5680 DAC 18 bit DAC, same level shift circuit and same control circuit. The reference is an LPRO-101. The Oscilloquartz ocxo was purchased used. Both UCT ocxo’s (only the better one is shown) were purchased new and have 100’s of operating hours. I have also seen differences with constant EFC control voltage. The differences limit what performance you can achieve.
TP
Tobias Pluess
Thu, Apr 2, 2020 7:13 PM

Hello all

in the meantime I figured out most of my problems and my GPSDO is working
now with some very ugly prototype code. Today, I wanted to do some ADEV
measurements.
My plan was to compare the 1PPS generated from my GPSDO to the 1PPS of my
Oscilloquartz STAR4; unfortunately I have nothing else (like Rb or so)
which is perhaps more stable. So I try with the STAR4 and see where I get.
However, before I did any meaningful measurements, I wanted to see what the
noise floor of my test equipment is.
Again, unfortunately I have nothing better than a HP 5335A with 1ns
resolution in TIC mode. I measured the noise floor of the TIC as follows:
the 1PPS output of my GPSDO was connected to a resistive power splitter,
and then, one output of the splitter went to channel A of the TIC (START
signal) while the other output from the splitter went first to a long cable
and then to channel B. With this, I achieved about 16ns of delay.
I then used the TIC together with Timelab and measured the ADEV of this
setup.
As far as I understand, if the delay of the cable stays constant (which it
does as long as it is not moved and the temperature stays the same), all I
see in the ADEV plot is the ADEV of my counter itself. Right?

So I let this test run for one hour (collected 3600 samples), and the
result looks terrible. See the attached file. I did the test twice; once I
used the STAR4 GPSDO as external reference for the counter, and once I used
its internal reference, which is a HP 10544A oven. As one can see, the ADEV
at 1sec is between 7e-10 and 8e-10. I don't know yet what numbers I can
expect from my GPSDO, but from datasheets of commercial GPSDOs I saw that
the ADEV shortly after powerup should be in the 1e-11 region. So how does
one measure such low ADEVs?

To me, it appears that the ADEV at 1sec is roughly the counter's
resolution; a bit less due to averaging. If I take averaging over 3600
samples into account, I think I could expect maybe ~1ns/sqrt(3600) =
16.7e-12 as ADEV at 1 second, but we can clearly see that this is not the
case. So there are two interesting questions arising:

a) I think the ADEV is so high because of the quantization error of the
counter. Assume the time interval measured is right at the transition from,
say, 15ns to 16ns, even the smallest amount of noise will produce some
alternating readings of 15ns and 16ns, which, in turn, results in an ADEV
around 1e-9, right? Further, why is this effect not averaged out with
sqrt(# of samples)?

b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution does
my counter need? If the above was true, I would expect that a 1ps
resolution (and an even better stability!) was required to measure ADEV of
1e-12, The fact that the (as far as I know) world's most recent,
rocket-science grade counter (some Keysight stuff) has "only" 20ps of
resolution, but people are still able to measure even 1e-14 shows that my
assumption is wrong. So how are the measurement resolution and the ADEV
related to each other? I plan to build my own TIC based on a TDC7200, which
would offer some 55ps of resolution, but how low could I go with that?

Best regards
Tobias
HB9FSX

On Fri, Mar 20, 2020 at 5:34 PM Bob Q bobqhome@live.com wrote:

I have seen differences between both UCT and Oscilloquartz 8663 ocxo’s.
The attached plot shows an example. Both boxes use Ublox LEA-6T receiver,
surveyed in, AD5680 DAC 18 bit DAC, same level shift circuit and same
control circuit. The reference is an LPRO-101. The Oscilloquartz ocxo was
purchased used. Both UCT ocxo’s (only the better one is shown) were
purchased new and have 100’s of operating hours. I have also seen
differences with constant EFC control voltage. The differences limit what
performance you can achieve._______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hello all in the meantime I figured out most of my problems and my GPSDO is working now with some very ugly prototype code. Today, I wanted to do some ADEV measurements. My plan was to compare the 1PPS generated from my GPSDO to the 1PPS of my Oscilloquartz STAR4; unfortunately I have nothing else (like Rb or so) which is perhaps more stable. So I try with the STAR4 and see where I get. However, before I did any meaningful measurements, I wanted to see what the noise floor of my test equipment is. Again, unfortunately I have nothing better than a HP 5335A with 1ns resolution in TIC mode. I measured the noise floor of the TIC as follows: the 1PPS output of my GPSDO was connected to a resistive power splitter, and then, one output of the splitter went to channel A of the TIC (START signal) while the other output from the splitter went first to a long cable and then to channel B. With this, I achieved about 16ns of delay. I then used the TIC together with Timelab and measured the ADEV of this setup. As far as I understand, if the delay of the cable stays constant (which it does as long as it is not moved and the temperature stays the same), all I see in the ADEV plot is the ADEV of my counter itself. Right? So I let this test run for one hour (collected 3600 samples), and the result looks terrible. See the attached file. I did the test twice; once I used the STAR4 GPSDO as external reference for the counter, and once I used its internal reference, which is a HP 10544A oven. As one can see, the ADEV at 1sec is between 7e-10 and 8e-10. I don't know yet what numbers I can expect from my GPSDO, but from datasheets of commercial GPSDOs I saw that the ADEV shortly after powerup should be in the 1e-11 region. So how does one measure such low ADEVs? To me, it appears that the ADEV at 1sec is roughly the counter's resolution; a bit less due to averaging. If I take averaging over 3600 samples into account, I think I could expect maybe ~1ns/sqrt(3600) = 16.7e-12 as ADEV at 1 second, but we can clearly see that this is not the case. So there are two interesting questions arising: a) I think the ADEV is so high because of the quantization error of the counter. Assume the time interval measured is right at the transition from, say, 15ns to 16ns, even the smallest amount of noise will produce some alternating readings of 15ns and 16ns, which, in turn, results in an ADEV around 1e-9, right? Further, why is this effect not averaged out with sqrt(# of samples)? b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution does my counter need? If the above was true, I would expect that a 1ps resolution (and an even better stability!) was required to measure ADEV of 1e-12, The fact that the (as far as I know) world's most recent, rocket-science grade counter (some Keysight stuff) has "only" 20ps of resolution, but people are still able to measure even 1e-14 shows that my assumption is wrong. So how are the measurement resolution and the ADEV related to each other? I plan to build my own TIC based on a TDC7200, which would offer some 55ps of resolution, but how low could I go with that? Best regards Tobias HB9FSX On Fri, Mar 20, 2020 at 5:34 PM Bob Q <bobqhome@live.com> wrote: > I have seen differences between both UCT and Oscilloquartz 8663 ocxo’s. > The attached plot shows an example. Both boxes use Ublox LEA-6T receiver, > surveyed in, AD5680 DAC 18 bit DAC, same level shift circuit and same > control circuit. The reference is an LPRO-101. The Oscilloquartz ocxo was > purchased used. Both UCT ocxo’s (only the better one is shown) were > purchased new and have 100’s of operating hours. I have also seen > differences with constant EFC control voltage. The differences limit what > performance you can achieve._______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >
BK
Bob kb8tq
Thu, Apr 2, 2020 8:17 PM

Hi

What you have measured is the noise floor of a 5335 when trying
to use it to measure ADEV. Anything past the numbers on your plot
will be “past” what the 5335 can “see”. Indeed, even when you get
close to those numbers, things may get a bit weird due to the fact
that you are measuring counter “noise” plus device noise.

Bob

On Apr 2, 2020, at 3:13 PM, Tobias Pluess tpluess@ieee.org wrote:

Hello all

in the meantime I figured out most of my problems and my GPSDO is working
now with some very ugly prototype code. Today, I wanted to do some ADEV
measurements.
My plan was to compare the 1PPS generated from my GPSDO to the 1PPS of my
Oscilloquartz STAR4; unfortunately I have nothing else (like Rb or so)
which is perhaps more stable. So I try with the STAR4 and see where I get.
However, before I did any meaningful measurements, I wanted to see what the
noise floor of my test equipment is.
Again, unfortunately I have nothing better than a HP 5335A with 1ns
resolution in TIC mode. I measured the noise floor of the TIC as follows:
the 1PPS output of my GPSDO was connected to a resistive power splitter,
and then, one output of the splitter went to channel A of the TIC (START
signal) while the other output from the splitter went first to a long cable
and then to channel B. With this, I achieved about 16ns of delay.
I then used the TIC together with Timelab and measured the ADEV of this
setup.
As far as I understand, if the delay of the cable stays constant (which it
does as long as it is not moved and the temperature stays the same), all I
see in the ADEV plot is the ADEV of my counter itself. Right?

So I let this test run for one hour (collected 3600 samples), and the
result looks terrible. See the attached file. I did the test twice; once I
used the STAR4 GPSDO as external reference for the counter, and once I used
its internal reference, which is a HP 10544A oven. As one can see, the ADEV
at 1sec is between 7e-10 and 8e-10. I don't know yet what numbers I can
expect from my GPSDO, but from datasheets of commercial GPSDOs I saw that
the ADEV shortly after powerup should be in the 1e-11 region. So how does
one measure such low ADEVs?

To me, it appears that the ADEV at 1sec is roughly the counter's
resolution; a bit less due to averaging. If I take averaging over 3600
samples into account, I think I could expect maybe ~1ns/sqrt(3600) =
16.7e-12 as ADEV at 1 second, but we can clearly see that this is not the
case. So there are two interesting questions arising:

a) I think the ADEV is so high because of the quantization error of the
counter. Assume the time interval measured is right at the transition from,
say, 15ns to 16ns, even the smallest amount of noise will produce some
alternating readings of 15ns and 16ns, which, in turn, results in an ADEV
around 1e-9, right? Further, why is this effect not averaged out with
sqrt(# of samples)?

b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution does
my counter need? If the above was true, I would expect that a 1ps
resolution (and an even better stability!) was required to measure ADEV of
1e-12, The fact that the (as far as I know) world's most recent,
rocket-science grade counter (some Keysight stuff) has "only" 20ps of
resolution, but people are still able to measure even 1e-14 shows that my
assumption is wrong. So how are the measurement resolution and the ADEV
related to each other? I plan to build my own TIC based on a TDC7200, which
would offer some 55ps of resolution, but how low could I go with that?

Best regards
Tobias
HB9FSX

On Fri, Mar 20, 2020 at 5:34 PM Bob Q bobqhome@live.com wrote:

I have seen differences between both UCT and Oscilloquartz 8663 ocxo’s.
The attached plot shows an example. Both boxes use Ublox LEA-6T receiver,
surveyed in, AD5680 DAC 18 bit DAC, same level shift circuit and same
control circuit. The reference is an LPRO-101. The Oscilloquartz ocxo was
purchased used. Both UCT ocxo’s (only the better one is shown) were
purchased new and have 100’s of operating hours. I have also seen
differences with constant EFC control voltage. The differences limit what
performance you can achieve._______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

<figure-1.png>_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi What you have measured *is* the noise floor of a 5335 when trying to use it to measure ADEV. Anything past the numbers on your plot will be “past” what the 5335 can “see”. Indeed, even when you get close to those numbers, things may get a bit weird due to the fact that you are measuring counter “noise” plus device noise. Bob > On Apr 2, 2020, at 3:13 PM, Tobias Pluess <tpluess@ieee.org> wrote: > > Hello all > > in the meantime I figured out most of my problems and my GPSDO is working > now with some very ugly prototype code. Today, I wanted to do some ADEV > measurements. > My plan was to compare the 1PPS generated from my GPSDO to the 1PPS of my > Oscilloquartz STAR4; unfortunately I have nothing else (like Rb or so) > which is perhaps more stable. So I try with the STAR4 and see where I get. > However, before I did any meaningful measurements, I wanted to see what the > noise floor of my test equipment is. > Again, unfortunately I have nothing better than a HP 5335A with 1ns > resolution in TIC mode. I measured the noise floor of the TIC as follows: > the 1PPS output of my GPSDO was connected to a resistive power splitter, > and then, one output of the splitter went to channel A of the TIC (START > signal) while the other output from the splitter went first to a long cable > and then to channel B. With this, I achieved about 16ns of delay. > I then used the TIC together with Timelab and measured the ADEV of this > setup. > As far as I understand, if the delay of the cable stays constant (which it > does as long as it is not moved and the temperature stays the same), all I > see in the ADEV plot is the ADEV of my counter itself. Right? > > So I let this test run for one hour (collected 3600 samples), and the > result looks terrible. See the attached file. I did the test twice; once I > used the STAR4 GPSDO as external reference for the counter, and once I used > its internal reference, which is a HP 10544A oven. As one can see, the ADEV > at 1sec is between 7e-10 and 8e-10. I don't know yet what numbers I can > expect from my GPSDO, but from datasheets of commercial GPSDOs I saw that > the ADEV shortly after powerup should be in the 1e-11 region. So how does > one measure such low ADEVs? > > To me, it appears that the ADEV at 1sec is roughly the counter's > resolution; a bit less due to averaging. If I take averaging over 3600 > samples into account, I think I could expect maybe ~1ns/sqrt(3600) = > 16.7e-12 as ADEV at 1 second, but we can clearly see that this is not the > case. So there are two interesting questions arising: > > a) I think the ADEV is so high because of the quantization error of the > counter. Assume the time interval measured is right at the transition from, > say, 15ns to 16ns, even the smallest amount of noise will produce some > alternating readings of 15ns and 16ns, which, in turn, results in an ADEV > around 1e-9, right? Further, why is this effect not averaged out with > sqrt(# of samples)? > > b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution does > my counter need? If the above was true, I would expect that a 1ps > resolution (and an even better stability!) was required to measure ADEV of > 1e-12, The fact that the (as far as I know) world's most recent, > rocket-science grade counter (some Keysight stuff) has "only" 20ps of > resolution, but people are still able to measure even 1e-14 shows that my > assumption is wrong. So how are the measurement resolution and the ADEV > related to each other? I plan to build my own TIC based on a TDC7200, which > would offer some 55ps of resolution, but how low could I go with that? > > > Best regards > Tobias > HB9FSX > > > > On Fri, Mar 20, 2020 at 5:34 PM Bob Q <bobqhome@live.com> wrote: > >> I have seen differences between both UCT and Oscilloquartz 8663 ocxo’s. >> The attached plot shows an example. Both boxes use Ublox LEA-6T receiver, >> surveyed in, AD5680 DAC 18 bit DAC, same level shift circuit and same >> control circuit. The reference is an LPRO-101. The Oscilloquartz ocxo was >> purchased used. Both UCT ocxo’s (only the better one is shown) were >> purchased new and have 100’s of operating hours. I have also seen >> differences with constant EFC control voltage. The differences limit what >> performance you can achieve._______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. >> > <figure-1.png>_______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
JM
John Miles
Thu, Apr 2, 2020 10:52 PM

b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution does
my counter need? If the above was true, I would expect that a 1ps
resolution (and an even better stability!) was required to measure ADEV of
1e-12, The fact that the (as far as I know) world's most recent,
rocket-science grade counter (some Keysight stuff) has "only" 20ps of
resolution, but people are still able to measure even 1e-14 shows that my
assumption is wrong. So how are the measurement resolution and the ADEV
related to each other? I plan to build my own TIC based on a TDC7200, which
would offer some 55ps of resolution, but how low could I go with that?

That sounds like a simple question but it's not.  There are a few different approaches to look into:

  1. Use averaging with your existing counter.  Some counters can yield readings in the 1E-12 region at t=1s even though their single-shot jitter is much worse than that.  They do this by averaging  hundreds or thousands of samples for each reading they report.  Whether (and when) this is acceptable is a complex topic in itself, too much so to explain quickly.  Search for information on the effects of averaging and dead time on Allan deviation to find the entrance to this fork of the rabbit hole.

  2. Search for the term 'DMTD' and read about that.

  3. Search for 'direct digital phase measurement' and read about that.

  4. Search for 'tight PLL' and read about that.

Basically, while some counters can perform averaging on a post-detection basis, that's like using the tone control on a radio to reduce static and QRM.  It works, sort of, but it's too late in the signal chain at that point to do the job right.  You really want to limit the bandwidth before the signal is captured, but since that's almost never practical at RF, the next best thing to do is limit the bandwidth before the signal is "demodulated" (i.e., counted.)

Hence items 2, 3, and 4 above.  They either limit the measurement bandwidth prior to detection, lower the frequency itself to keep the counter's inherent jitter from dominating the measurement, or both.  You'll have to use one of these methods, or another technique along the same lines, if you want to measure the short-term stability of a good oscillator or GPSDO.

-- john, KE5FX

> b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution does > my counter need? If the above was true, I would expect that a 1ps > resolution (and an even better stability!) was required to measure ADEV of > 1e-12, The fact that the (as far as I know) world's most recent, > rocket-science grade counter (some Keysight stuff) has "only" 20ps of > resolution, but people are still able to measure even 1e-14 shows that my > assumption is wrong. So how are the measurement resolution and the ADEV > related to each other? I plan to build my own TIC based on a TDC7200, which > would offer some 55ps of resolution, but how low could I go with that? That sounds like a simple question but it's not. There are a few different approaches to look into: 1) Use averaging with your existing counter. Some counters can yield readings in the 1E-12 region at t=1s even though their single-shot jitter is much worse than that. They do this by averaging hundreds or thousands of samples for each reading they report. Whether (and when) this is acceptable is a complex topic in itself, too much so to explain quickly. Search for information on the effects of averaging and dead time on Allan deviation to find the entrance to this fork of the rabbit hole. 2) Search for the term 'DMTD' and read about that. 3) Search for 'direct digital phase measurement' and read about that. 4) Search for 'tight PLL' and read about that. Basically, while some counters can perform averaging on a post-detection basis, that's like using the tone control on a radio to reduce static and QRM. It works, sort of, but it's too late in the signal chain at that point to do the job right. You really want to limit the bandwidth before the signal is captured, but since that's almost never practical at RF, the next best thing to do is limit the bandwidth before the signal is "demodulated" (i.e., counted.) Hence items 2, 3, and 4 above. They either limit the measurement bandwidth prior to detection, lower the frequency itself to keep the counter's inherent jitter from dominating the measurement, or both. You'll have to use one of these methods, or another technique along the same lines, if you want to measure the short-term stability of a good oscillator or GPSDO. -- john, KE5FX
TP
Tobias Pluess
Fri, Apr 3, 2020 9:13 AM

hi John

yes I know the DMTD method, and indeed I am planing to build my own DMTD
system, something similar to the "Small DMTD system" published by Riley (
https://www.wriley.com/A Small DMTD System.pdf).
However I am unsure whether that will help much in this case, because all
what the DMTD does is to mix the 10MHz signals down to some 1Hz Signal or
so which can be measured more easily, and I already have 1Hz signals (the
1PPS) which I am comparing.
Or do you suggest to use the DMTD and use a higher frequency at its
outputs, say 10Hz or so, and then average for 10 samples  to increase the
resolution?

Thanks
Tobias
HB9FSX

On Fri, Apr 3, 2020 at 12:53 AM John Miles john@miles.io wrote:

b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution

does

my counter need? If the above was true, I would expect that a 1ps
resolution (and an even better stability!) was required to measure ADEV

of

1e-12, The fact that the (as far as I know) world's most recent,
rocket-science grade counter (some Keysight stuff) has "only" 20ps of
resolution, but people are still able to measure even 1e-14 shows that my
assumption is wrong. So how are the measurement resolution and the ADEV
related to each other? I plan to build my own TIC based on a TDC7200,

which

would offer some 55ps of resolution, but how low could I go with that?

That sounds like a simple question but it's not.  There are a few
different approaches to look into:

  1. Use averaging with your existing counter.  Some counters can yield
    readings in the 1E-12 region at t=1s even though their single-shot jitter
    is much worse than that.  They do this by averaging  hundreds or thousands
    of samples for each reading they report.  Whether (and when) this is
    acceptable is a complex topic in itself, too much so to explain quickly.
    Search for information on the effects of averaging and dead time on Allan
    deviation to find the entrance to this fork of the rabbit hole.

  2. Search for the term 'DMTD' and read about that.

  3. Search for 'direct digital phase measurement' and read about that.

  4. Search for 'tight PLL' and read about that.

Basically, while some counters can perform averaging on a post-detection
basis, that's like using the tone control on a radio to reduce static and
QRM.  It works, sort of, but it's too late in the signal chain at that
point to do the job right.  You really want to limit the bandwidth before
the signal is captured, but since that's almost never practical at RF, the
next best thing to do is limit the bandwidth before the signal is
"demodulated" (i.e., counted.)

Hence items 2, 3, and 4 above.  They either limit the measurement
bandwidth prior to detection, lower the frequency itself to keep the
counter's inherent jitter from dominating the measurement, or both.  You'll
have to use one of these methods, or another technique along the same
lines, if you want to measure the short-term stability of a good oscillator
or GPSDO.

-- john, KE5FX


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

hi John yes I know the DMTD method, and indeed I am planing to build my own DMTD system, something similar to the "Small DMTD system" published by Riley ( https://www.wriley.com/A Small DMTD System.pdf). However I am unsure whether that will help much in this case, because all what the DMTD does is to mix the 10MHz signals down to some 1Hz Signal or so which can be measured more easily, and I already have 1Hz signals (the 1PPS) which I am comparing. Or do you suggest to use the DMTD and use a higher frequency at its outputs, say 10Hz or so, and then average for 10 samples to increase the resolution? Thanks Tobias HB9FSX On Fri, Apr 3, 2020 at 12:53 AM John Miles <john@miles.io> wrote: > > b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution > does > > my counter need? If the above was true, I would expect that a 1ps > > resolution (and an even better stability!) was required to measure ADEV > of > > 1e-12, The fact that the (as far as I know) world's most recent, > > rocket-science grade counter (some Keysight stuff) has "only" 20ps of > > resolution, but people are still able to measure even 1e-14 shows that my > > assumption is wrong. So how are the measurement resolution and the ADEV > > related to each other? I plan to build my own TIC based on a TDC7200, > which > > would offer some 55ps of resolution, but how low could I go with that? > > That sounds like a simple question but it's not. There are a few > different approaches to look into: > > 1) Use averaging with your existing counter. Some counters can yield > readings in the 1E-12 region at t=1s even though their single-shot jitter > is much worse than that. They do this by averaging hundreds or thousands > of samples for each reading they report. Whether (and when) this is > acceptable is a complex topic in itself, too much so to explain quickly. > Search for information on the effects of averaging and dead time on Allan > deviation to find the entrance to this fork of the rabbit hole. > > 2) Search for the term 'DMTD' and read about that. > > 3) Search for 'direct digital phase measurement' and read about that. > > 4) Search for 'tight PLL' and read about that. > > Basically, while some counters can perform averaging on a post-detection > basis, that's like using the tone control on a radio to reduce static and > QRM. It works, sort of, but it's too late in the signal chain at that > point to do the job right. You really want to limit the bandwidth before > the signal is captured, but since that's almost never practical at RF, the > next best thing to do is limit the bandwidth before the signal is > "demodulated" (i.e., counted.) > > Hence items 2, 3, and 4 above. They either limit the measurement > bandwidth prior to detection, lower the frequency itself to keep the > counter's inherent jitter from dominating the measurement, or both. You'll > have to use one of these methods, or another technique along the same > lines, if you want to measure the short-term stability of a good oscillator > or GPSDO. > > -- john, KE5FX > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >
BK
Bob kb8tq
Fri, Apr 3, 2020 11:44 AM

Hi

The quick way to do this is with a single mixer. Take something like an old
10811 and use the coarse tune to set it high in frequency by 5 to 10 Hz.

Then feed it into an RPD-1 mixer and pull out the 5 to 10 Hz audio tone.
That tone is the difference between the 10811 and your device under test.
If the DUT moves 1 Hz, the audio tone changes by 1 Hz.

If you measured the 10 MHz on the DUT, that 1 Hz would be a very small shift
( 0.1 ppm ). At 10 Hz it’s a 10% change. You have “amplified” the change
in frequency by the ratio of 10 MHz to 10 Hz ( so a million X increase ).

IF you could tack that on to the ADEV plot of your 5335 ( no, it’s not that
simple) your 7x10^-10 at 1 second would become more 7x10^-16 at 1
second.

The reason its not quite that simple is that the input circuit on the counter
really does not handle a 10 Hz audio tone as well as it handles a 10 MHz
RF signal. Instead of getting 9 digits a second, you probably will get three
good digits a second and another 6 digits of noise.

The good news is that an op amp used as a preamp ( to get you up to maybe
32 V p-p rather than a volt or so) and another op amp or three as limiters will
get you up around 6 or 7 good digits. Toss in a cap or two as a high pass
and low pass filter ( DC offsets can be a problem ….) and you have a working
device that gets into the parts in 10^-13 with your 5335.

It all can be done with point to point wiring. No need for a PCB layout. Be
careful that the +/- 18V supplies to the op amp both go on and off at the
same time ….

Bob

On Apr 3, 2020, at 5:13 AM, Tobias Pluess tpluess@ieee.org wrote:

hi John

yes I know the DMTD method, and indeed I am planing to build my own DMTD
system, something similar to the "Small DMTD system" published by Riley (
https://www.wriley.com/A Small DMTD System.pdf).
However I am unsure whether that will help much in this case, because all
what the DMTD does is to mix the 10MHz signals down to some 1Hz Signal or
so which can be measured more easily, and I already have 1Hz signals (the
1PPS) which I am comparing.
Or do you suggest to use the DMTD and use a higher frequency at its
outputs, say 10Hz or so, and then average for 10 samples  to increase the
resolution?

Thanks
Tobias
HB9FSX

On Fri, Apr 3, 2020 at 12:53 AM John Miles john@miles.io wrote:

b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution

does

my counter need? If the above was true, I would expect that a 1ps
resolution (and an even better stability!) was required to measure ADEV

of

1e-12, The fact that the (as far as I know) world's most recent,
rocket-science grade counter (some Keysight stuff) has "only" 20ps of
resolution, but people are still able to measure even 1e-14 shows that my
assumption is wrong. So how are the measurement resolution and the ADEV
related to each other? I plan to build my own TIC based on a TDC7200,

which

would offer some 55ps of resolution, but how low could I go with that?

That sounds like a simple question but it's not.  There are a few
different approaches to look into:

  1. Use averaging with your existing counter.  Some counters can yield
    readings in the 1E-12 region at t=1s even though their single-shot jitter
    is much worse than that.  They do this by averaging  hundreds or thousands
    of samples for each reading they report.  Whether (and when) this is
    acceptable is a complex topic in itself, too much so to explain quickly.
    Search for information on the effects of averaging and dead time on Allan
    deviation to find the entrance to this fork of the rabbit hole.

  2. Search for the term 'DMTD' and read about that.

  3. Search for 'direct digital phase measurement' and read about that.

  4. Search for 'tight PLL' and read about that.

Basically, while some counters can perform averaging on a post-detection
basis, that's like using the tone control on a radio to reduce static and
QRM.  It works, sort of, but it's too late in the signal chain at that
point to do the job right.  You really want to limit the bandwidth before
the signal is captured, but since that's almost never practical at RF, the
next best thing to do is limit the bandwidth before the signal is
"demodulated" (i.e., counted.)

Hence items 2, 3, and 4 above.  They either limit the measurement
bandwidth prior to detection, lower the frequency itself to keep the
counter's inherent jitter from dominating the measurement, or both.  You'll
have to use one of these methods, or another technique along the same
lines, if you want to measure the short-term stability of a good oscillator
or GPSDO.

-- john, KE5FX


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi The quick way to do this is with a single mixer. Take something like an old 10811 and use the coarse tune to set it high in frequency by 5 to 10 Hz. Then feed it into an RPD-1 mixer and pull out the 5 to 10 Hz audio tone. That tone is the *difference* between the 10811 and your device under test. If the DUT moves 1 Hz, the audio tone changes by 1 Hz. If you measured the 10 MHz on the DUT, that 1 Hz would be a very small shift ( 0.1 ppm ). At 10 Hz it’s a 10% change. You have “amplified” the change in frequency by the ratio of 10 MHz to 10 Hz ( so a million X increase ). *IF* you could tack that on to the ADEV plot of your 5335 ( no, it’s not that simple) your 7x10^-10 at 1 second would become more 7x10^-16 at 1 second. The reason its not quite that simple is that the input circuit on the counter really does not handle a 10 Hz audio tone as well as it handles a 10 MHz RF signal. Instead of getting 9 digits a second, you probably will get three *good* digits a second and another 6 digits of noise. The good news is that an op amp used as a preamp ( to get you up to maybe 32 V p-p rather than a volt or so) and another op amp or three as limiters will get you up around 6 or 7 good digits. Toss in a cap or two as a high pass and low pass filter ( DC offsets can be a problem ….) and you have a working device that gets into the parts in 10^-13 with your 5335. It all can be done with point to point wiring. No need for a PCB layout. Be careful that the +/- 18V supplies to the op amp *both* go on and off at the same time …. Bob > On Apr 3, 2020, at 5:13 AM, Tobias Pluess <tpluess@ieee.org> wrote: > > hi John > > yes I know the DMTD method, and indeed I am planing to build my own DMTD > system, something similar to the "Small DMTD system" published by Riley ( > https://www.wriley.com/A Small DMTD System.pdf). > However I am unsure whether that will help much in this case, because all > what the DMTD does is to mix the 10MHz signals down to some 1Hz Signal or > so which can be measured more easily, and I already have 1Hz signals (the > 1PPS) which I am comparing. > Or do you suggest to use the DMTD and use a higher frequency at its > outputs, say 10Hz or so, and then average for 10 samples to increase the > resolution? > > Thanks > Tobias > HB9FSX > > > On Fri, Apr 3, 2020 at 12:53 AM John Miles <john@miles.io> wrote: > >>> b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution >> does >>> my counter need? If the above was true, I would expect that a 1ps >>> resolution (and an even better stability!) was required to measure ADEV >> of >>> 1e-12, The fact that the (as far as I know) world's most recent, >>> rocket-science grade counter (some Keysight stuff) has "only" 20ps of >>> resolution, but people are still able to measure even 1e-14 shows that my >>> assumption is wrong. So how are the measurement resolution and the ADEV >>> related to each other? I plan to build my own TIC based on a TDC7200, >> which >>> would offer some 55ps of resolution, but how low could I go with that? >> >> That sounds like a simple question but it's not. There are a few >> different approaches to look into: >> >> 1) Use averaging with your existing counter. Some counters can yield >> readings in the 1E-12 region at t=1s even though their single-shot jitter >> is much worse than that. They do this by averaging hundreds or thousands >> of samples for each reading they report. Whether (and when) this is >> acceptable is a complex topic in itself, too much so to explain quickly. >> Search for information on the effects of averaging and dead time on Allan >> deviation to find the entrance to this fork of the rabbit hole. >> >> 2) Search for the term 'DMTD' and read about that. >> >> 3) Search for 'direct digital phase measurement' and read about that. >> >> 4) Search for 'tight PLL' and read about that. >> >> Basically, while some counters can perform averaging on a post-detection >> basis, that's like using the tone control on a radio to reduce static and >> QRM. It works, sort of, but it's too late in the signal chain at that >> point to do the job right. You really want to limit the bandwidth before >> the signal is captured, but since that's almost never practical at RF, the >> next best thing to do is limit the bandwidth before the signal is >> "demodulated" (i.e., counted.) >> >> Hence items 2, 3, and 4 above. They either limit the measurement >> bandwidth prior to detection, lower the frequency itself to keep the >> counter's inherent jitter from dominating the measurement, or both. You'll >> have to use one of these methods, or another technique along the same >> lines, if you want to measure the short-term stability of a good oscillator >> or GPSDO. >> >> -- john, KE5FX >> >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
DW
Dana Whitlow
Fri, Apr 3, 2020 1:29 PM

Caution: opamps make terrible limiters- their overload behavior is
generally ugly
and unpredictable.  It's much better to use a genuine level comparator, and
wire it
up so that it has a modest amount of hysteresis.

Dana

On Fri, Apr 3, 2020 at 6:45 AM Bob kb8tq kb8tq@n1k.org wrote:

Hi

The quick way to do this is with a single mixer. Take something like an old
10811 and use the coarse tune to set it high in frequency by 5 to 10 Hz.

Then feed it into an RPD-1 mixer and pull out the 5 to 10 Hz audio tone.
That tone is the difference between the 10811 and your device under
test.
If the DUT moves 1 Hz, the audio tone changes by 1 Hz.

If you measured the 10 MHz on the DUT, that 1 Hz would be a very small
shift
( 0.1 ppm ). At 10 Hz it’s a 10% change. You have “amplified” the change
in frequency by the ratio of 10 MHz to 10 Hz ( so a million X increase ).

IF you could tack that on to the ADEV plot of your 5335 ( no, it’s not
that
simple) your 7x10^-10 at 1 second would become more 7x10^-16 at 1
second.

The reason its not quite that simple is that the input circuit on the
counter
really does not handle a 10 Hz audio tone as well as it handles a 10 MHz
RF signal. Instead of getting 9 digits a second, you probably will get
three
good digits a second and another 6 digits of noise.

The good news is that an op amp used as a preamp ( to get you up to maybe
32 V p-p rather than a volt or so) and another op amp or three as limiters
will
get you up around 6 or 7 good digits. Toss in a cap or two as a high pass
and low pass filter ( DC offsets can be a problem ….) and you have a
working
device that gets into the parts in 10^-13 with your 5335.

It all can be done with point to point wiring. No need for a PCB layout.
Be
careful that the +/- 18V supplies to the op amp both go on and off at
the
same time ….

Bob

On Apr 3, 2020, at 5:13 AM, Tobias Pluess tpluess@ieee.org wrote:

hi John

yes I know the DMTD method, and indeed I am planing to build my own DMTD
system, something similar to the "Small DMTD system" published by Riley (
https://www.wriley.com/A Small DMTD System.pdf).
However I am unsure whether that will help much in this case, because all
what the DMTD does is to mix the 10MHz signals down to some 1Hz Signal or
so which can be measured more easily, and I already have 1Hz signals (the
1PPS) which I am comparing.
Or do you suggest to use the DMTD and use a higher frequency at its
outputs, say 10Hz or so, and then average for 10 samples  to increase the
resolution?

Thanks
Tobias
HB9FSX

On Fri, Apr 3, 2020 at 12:53 AM John Miles john@miles.io wrote:

b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution

does

my counter need? If the above was true, I would expect that a 1ps
resolution (and an even better stability!) was required to measure ADEV

of

1e-12, The fact that the (as far as I know) world's most recent,
rocket-science grade counter (some Keysight stuff) has "only" 20ps of
resolution, but people are still able to measure even 1e-14 shows that

my

assumption is wrong. So how are the measurement resolution and the ADEV
related to each other? I plan to build my own TIC based on a TDC7200,

which

would offer some 55ps of resolution, but how low could I go with that?

That sounds like a simple question but it's not.  There are a few
different approaches to look into:

  1. Use averaging with your existing counter.  Some counters can yield
    readings in the 1E-12 region at t=1s even though their single-shot

jitter

is much worse than that.  They do this by averaging  hundreds or

thousands

of samples for each reading they report.  Whether (and when) this is
acceptable is a complex topic in itself, too much so to explain quickly.
Search for information on the effects of averaging and dead time on

Allan

deviation to find the entrance to this fork of the rabbit hole.

  1. Search for the term 'DMTD' and read about that.

  2. Search for 'direct digital phase measurement' and read about that.

  3. Search for 'tight PLL' and read about that.

Basically, while some counters can perform averaging on a post-detection
basis, that's like using the tone control on a radio to reduce static

and

QRM.  It works, sort of, but it's too late in the signal chain at that
point to do the job right.  You really want to limit the bandwidth

before

the signal is captured, but since that's almost never practical at RF,

the

next best thing to do is limit the bandwidth before the signal is
"demodulated" (i.e., counted.)

Hence items 2, 3, and 4 above.  They either limit the measurement
bandwidth prior to detection, lower the frequency itself to keep the
counter's inherent jitter from dominating the measurement, or both.

You'll

have to use one of these methods, or another technique along the same
lines, if you want to measure the short-term stability of a good

oscillator

or GPSDO.

-- john, KE5FX


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Caution: opamps make terrible limiters- their overload behavior is generally ugly and unpredictable. It's much better to use a genuine level comparator, and wire it up so that it has a modest amount of hysteresis. Dana On Fri, Apr 3, 2020 at 6:45 AM Bob kb8tq <kb8tq@n1k.org> wrote: > Hi > > The quick way to do this is with a single mixer. Take something like an old > 10811 and use the coarse tune to set it high in frequency by 5 to 10 Hz. > > Then feed it into an RPD-1 mixer and pull out the 5 to 10 Hz audio tone. > That tone is the *difference* between the 10811 and your device under > test. > If the DUT moves 1 Hz, the audio tone changes by 1 Hz. > > If you measured the 10 MHz on the DUT, that 1 Hz would be a very small > shift > ( 0.1 ppm ). At 10 Hz it’s a 10% change. You have “amplified” the change > in frequency by the ratio of 10 MHz to 10 Hz ( so a million X increase ). > > *IF* you could tack that on to the ADEV plot of your 5335 ( no, it’s not > that > simple) your 7x10^-10 at 1 second would become more 7x10^-16 at 1 > second. > > The reason its not quite that simple is that the input circuit on the > counter > really does not handle a 10 Hz audio tone as well as it handles a 10 MHz > RF signal. Instead of getting 9 digits a second, you probably will get > three > *good* digits a second and another 6 digits of noise. > > The good news is that an op amp used as a preamp ( to get you up to maybe > 32 V p-p rather than a volt or so) and another op amp or three as limiters > will > get you up around 6 or 7 good digits. Toss in a cap or two as a high pass > and low pass filter ( DC offsets can be a problem ….) and you have a > working > device that gets into the parts in 10^-13 with your 5335. > > It all can be done with point to point wiring. No need for a PCB layout. > Be > careful that the +/- 18V supplies to the op amp *both* go on and off at > the > same time …. > > Bob > > > On Apr 3, 2020, at 5:13 AM, Tobias Pluess <tpluess@ieee.org> wrote: > > > > hi John > > > > yes I know the DMTD method, and indeed I am planing to build my own DMTD > > system, something similar to the "Small DMTD system" published by Riley ( > > https://www.wriley.com/A Small DMTD System.pdf). > > However I am unsure whether that will help much in this case, because all > > what the DMTD does is to mix the 10MHz signals down to some 1Hz Signal or > > so which can be measured more easily, and I already have 1Hz signals (the > > 1PPS) which I am comparing. > > Or do you suggest to use the DMTD and use a higher frequency at its > > outputs, say 10Hz or so, and then average for 10 samples to increase the > > resolution? > > > > Thanks > > Tobias > > HB9FSX > > > > > > On Fri, Apr 3, 2020 at 12:53 AM John Miles <john@miles.io> wrote: > > > >>> b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution > >> does > >>> my counter need? If the above was true, I would expect that a 1ps > >>> resolution (and an even better stability!) was required to measure ADEV > >> of > >>> 1e-12, The fact that the (as far as I know) world's most recent, > >>> rocket-science grade counter (some Keysight stuff) has "only" 20ps of > >>> resolution, but people are still able to measure even 1e-14 shows that > my > >>> assumption is wrong. So how are the measurement resolution and the ADEV > >>> related to each other? I plan to build my own TIC based on a TDC7200, > >> which > >>> would offer some 55ps of resolution, but how low could I go with that? > >> > >> That sounds like a simple question but it's not. There are a few > >> different approaches to look into: > >> > >> 1) Use averaging with your existing counter. Some counters can yield > >> readings in the 1E-12 region at t=1s even though their single-shot > jitter > >> is much worse than that. They do this by averaging hundreds or > thousands > >> of samples for each reading they report. Whether (and when) this is > >> acceptable is a complex topic in itself, too much so to explain quickly. > >> Search for information on the effects of averaging and dead time on > Allan > >> deviation to find the entrance to this fork of the rabbit hole. > >> > >> 2) Search for the term 'DMTD' and read about that. > >> > >> 3) Search for 'direct digital phase measurement' and read about that. > >> > >> 4) Search for 'tight PLL' and read about that. > >> > >> Basically, while some counters can perform averaging on a post-detection > >> basis, that's like using the tone control on a radio to reduce static > and > >> QRM. It works, sort of, but it's too late in the signal chain at that > >> point to do the job right. You really want to limit the bandwidth > before > >> the signal is captured, but since that's almost never practical at RF, > the > >> next best thing to do is limit the bandwidth before the signal is > >> "demodulated" (i.e., counted.) > >> > >> Hence items 2, 3, and 4 above. They either limit the measurement > >> bandwidth prior to detection, lower the frequency itself to keep the > >> counter's inherent jitter from dominating the measurement, or both. > You'll > >> have to use one of these methods, or another technique along the same > >> lines, if you want to measure the short-term stability of a good > oscillator > >> or GPSDO. > >> > >> -- john, KE5FX > >> > >> > >> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@lists.febo.com > >> To unsubscribe, go to > >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >> and follow the instructions there. > >> > > _______________________________________________ > > time-nuts mailing list -- time-nuts@lists.febo.com > > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >
BK
Bob kb8tq
Fri, Apr 3, 2020 2:55 PM

Hi

If you are doing a limiter for a 5 Hz sine wave, a “normal” comparator is very much
not what you want to use. The slew rates involved are simply way to far below what
it is targeted to do. Effectively, it’s what’s in the counter input circuit that has already
failed miserably trying to do this.  Think of this as a “slow sine to square converter”.

For all the ugly details the discussions a few years back on the Colins style limiter
get into the this and that. Bruce did a nice write up at:

http://www.ko4bb.com/getsimple/index.php?id=bruces-zero-crossing-detectors

Bob

On Apr 3, 2020, at 9:29 AM, Dana Whitlow k8yumdoober@gmail.com wrote:

Caution: opamps make terrible limiters- their overload behavior is
generally ugly
and unpredictable.  It's much better to use a genuine level comparator, and
wire it
up so that it has a modest amount of hysteresis.

Dana

On Fri, Apr 3, 2020 at 6:45 AM Bob kb8tq kb8tq@n1k.org wrote:

Hi

The quick way to do this is with a single mixer. Take something like an old
10811 and use the coarse tune to set it high in frequency by 5 to 10 Hz.

Then feed it into an RPD-1 mixer and pull out the 5 to 10 Hz audio tone.
That tone is the difference between the 10811 and your device under
test.
If the DUT moves 1 Hz, the audio tone changes by 1 Hz.

If you measured the 10 MHz on the DUT, that 1 Hz would be a very small
shift
( 0.1 ppm ). At 10 Hz it’s a 10% change. You have “amplified” the change
in frequency by the ratio of 10 MHz to 10 Hz ( so a million X increase ).

IF you could tack that on to the ADEV plot of your 5335 ( no, it’s not
that
simple) your 7x10^-10 at 1 second would become more 7x10^-16 at 1
second.

The reason its not quite that simple is that the input circuit on the
counter
really does not handle a 10 Hz audio tone as well as it handles a 10 MHz
RF signal. Instead of getting 9 digits a second, you probably will get
three
good digits a second and another 6 digits of noise.

The good news is that an op amp used as a preamp ( to get you up to maybe
32 V p-p rather than a volt or so) and another op amp or three as limiters
will
get you up around 6 or 7 good digits. Toss in a cap or two as a high pass
and low pass filter ( DC offsets can be a problem ….) and you have a
working
device that gets into the parts in 10^-13 with your 5335.

It all can be done with point to point wiring. No need for a PCB layout.
Be
careful that the +/- 18V supplies to the op amp both go on and off at
the
same time ….

Bob

On Apr 3, 2020, at 5:13 AM, Tobias Pluess tpluess@ieee.org wrote:

hi John

yes I know the DMTD method, and indeed I am planing to build my own DMTD
system, something similar to the "Small DMTD system" published by Riley (
https://www.wriley.com/A Small DMTD System.pdf).
However I am unsure whether that will help much in this case, because all
what the DMTD does is to mix the 10MHz signals down to some 1Hz Signal or
so which can be measured more easily, and I already have 1Hz signals (the
1PPS) which I am comparing.
Or do you suggest to use the DMTD and use a higher frequency at its
outputs, say 10Hz or so, and then average for 10 samples  to increase the
resolution?

Thanks
Tobias
HB9FSX

On Fri, Apr 3, 2020 at 12:53 AM John Miles john@miles.io wrote:

b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution

does

my counter need? If the above was true, I would expect that a 1ps
resolution (and an even better stability!) was required to measure ADEV

of

1e-12, The fact that the (as far as I know) world's most recent,
rocket-science grade counter (some Keysight stuff) has "only" 20ps of
resolution, but people are still able to measure even 1e-14 shows that

my

assumption is wrong. So how are the measurement resolution and the ADEV
related to each other? I plan to build my own TIC based on a TDC7200,

which

would offer some 55ps of resolution, but how low could I go with that?

That sounds like a simple question but it's not.  There are a few
different approaches to look into:

  1. Use averaging with your existing counter.  Some counters can yield
    readings in the 1E-12 region at t=1s even though their single-shot

jitter

is much worse than that.  They do this by averaging  hundreds or

thousands

of samples for each reading they report.  Whether (and when) this is
acceptable is a complex topic in itself, too much so to explain quickly.
Search for information on the effects of averaging and dead time on

Allan

deviation to find the entrance to this fork of the rabbit hole.

  1. Search for the term 'DMTD' and read about that.

  2. Search for 'direct digital phase measurement' and read about that.

  3. Search for 'tight PLL' and read about that.

Basically, while some counters can perform averaging on a post-detection
basis, that's like using the tone control on a radio to reduce static

and

QRM.  It works, sort of, but it's too late in the signal chain at that
point to do the job right.  You really want to limit the bandwidth

before

the signal is captured, but since that's almost never practical at RF,

the

next best thing to do is limit the bandwidth before the signal is
"demodulated" (i.e., counted.)

Hence items 2, 3, and 4 above.  They either limit the measurement
bandwidth prior to detection, lower the frequency itself to keep the
counter's inherent jitter from dominating the measurement, or both.

You'll

have to use one of these methods, or another technique along the same
lines, if you want to measure the short-term stability of a good

oscillator

or GPSDO.

-- john, KE5FX


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi If you are doing a limiter for a 5 Hz sine wave, a “normal” comparator is very much not what you want to use. The slew rates involved are simply way to far below what it is targeted to do. Effectively, it’s what’s in the counter input circuit that has already failed miserably trying to do this. Think of this as a “slow sine to square converter”. For all the ugly details the discussions a few years back on the Colins style limiter get into the this and that. Bruce did a nice write up at: http://www.ko4bb.com/getsimple/index.php?id=bruces-zero-crossing-detectors Bob > On Apr 3, 2020, at 9:29 AM, Dana Whitlow <k8yumdoober@gmail.com> wrote: > > Caution: opamps make terrible limiters- their overload behavior is > generally ugly > and unpredictable. It's much better to use a genuine level comparator, and > wire it > up so that it has a modest amount of hysteresis. > > Dana > > > On Fri, Apr 3, 2020 at 6:45 AM Bob kb8tq <kb8tq@n1k.org> wrote: > >> Hi >> >> The quick way to do this is with a single mixer. Take something like an old >> 10811 and use the coarse tune to set it high in frequency by 5 to 10 Hz. >> >> Then feed it into an RPD-1 mixer and pull out the 5 to 10 Hz audio tone. >> That tone is the *difference* between the 10811 and your device under >> test. >> If the DUT moves 1 Hz, the audio tone changes by 1 Hz. >> >> If you measured the 10 MHz on the DUT, that 1 Hz would be a very small >> shift >> ( 0.1 ppm ). At 10 Hz it’s a 10% change. You have “amplified” the change >> in frequency by the ratio of 10 MHz to 10 Hz ( so a million X increase ). >> >> *IF* you could tack that on to the ADEV plot of your 5335 ( no, it’s not >> that >> simple) your 7x10^-10 at 1 second would become more 7x10^-16 at 1 >> second. >> >> The reason its not quite that simple is that the input circuit on the >> counter >> really does not handle a 10 Hz audio tone as well as it handles a 10 MHz >> RF signal. Instead of getting 9 digits a second, you probably will get >> three >> *good* digits a second and another 6 digits of noise. >> >> The good news is that an op amp used as a preamp ( to get you up to maybe >> 32 V p-p rather than a volt or so) and another op amp or three as limiters >> will >> get you up around 6 or 7 good digits. Toss in a cap or two as a high pass >> and low pass filter ( DC offsets can be a problem ….) and you have a >> working >> device that gets into the parts in 10^-13 with your 5335. >> >> It all can be done with point to point wiring. No need for a PCB layout. >> Be >> careful that the +/- 18V supplies to the op amp *both* go on and off at >> the >> same time …. >> >> Bob >> >>> On Apr 3, 2020, at 5:13 AM, Tobias Pluess <tpluess@ieee.org> wrote: >>> >>> hi John >>> >>> yes I know the DMTD method, and indeed I am planing to build my own DMTD >>> system, something similar to the "Small DMTD system" published by Riley ( >>> https://www.wriley.com/A Small DMTD System.pdf). >>> However I am unsure whether that will help much in this case, because all >>> what the DMTD does is to mix the 10MHz signals down to some 1Hz Signal or >>> so which can be measured more easily, and I already have 1Hz signals (the >>> 1PPS) which I am comparing. >>> Or do you suggest to use the DMTD and use a higher frequency at its >>> outputs, say 10Hz or so, and then average for 10 samples to increase the >>> resolution? >>> >>> Thanks >>> Tobias >>> HB9FSX >>> >>> >>> On Fri, Apr 3, 2020 at 12:53 AM John Miles <john@miles.io> wrote: >>> >>>>> b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution >>>> does >>>>> my counter need? If the above was true, I would expect that a 1ps >>>>> resolution (and an even better stability!) was required to measure ADEV >>>> of >>>>> 1e-12, The fact that the (as far as I know) world's most recent, >>>>> rocket-science grade counter (some Keysight stuff) has "only" 20ps of >>>>> resolution, but people are still able to measure even 1e-14 shows that >> my >>>>> assumption is wrong. So how are the measurement resolution and the ADEV >>>>> related to each other? I plan to build my own TIC based on a TDC7200, >>>> which >>>>> would offer some 55ps of resolution, but how low could I go with that? >>>> >>>> That sounds like a simple question but it's not. There are a few >>>> different approaches to look into: >>>> >>>> 1) Use averaging with your existing counter. Some counters can yield >>>> readings in the 1E-12 region at t=1s even though their single-shot >> jitter >>>> is much worse than that. They do this by averaging hundreds or >> thousands >>>> of samples for each reading they report. Whether (and when) this is >>>> acceptable is a complex topic in itself, too much so to explain quickly. >>>> Search for information on the effects of averaging and dead time on >> Allan >>>> deviation to find the entrance to this fork of the rabbit hole. >>>> >>>> 2) Search for the term 'DMTD' and read about that. >>>> >>>> 3) Search for 'direct digital phase measurement' and read about that. >>>> >>>> 4) Search for 'tight PLL' and read about that. >>>> >>>> Basically, while some counters can perform averaging on a post-detection >>>> basis, that's like using the tone control on a radio to reduce static >> and >>>> QRM. It works, sort of, but it's too late in the signal chain at that >>>> point to do the job right. You really want to limit the bandwidth >> before >>>> the signal is captured, but since that's almost never practical at RF, >> the >>>> next best thing to do is limit the bandwidth before the signal is >>>> "demodulated" (i.e., counted.) >>>> >>>> Hence items 2, 3, and 4 above. They either limit the measurement >>>> bandwidth prior to detection, lower the frequency itself to keep the >>>> counter's inherent jitter from dominating the measurement, or both. >> You'll >>>> have to use one of these methods, or another technique along the same >>>> lines, if you want to measure the short-term stability of a good >> oscillator >>>> or GPSDO. >>>> >>>> -- john, KE5FX >>>> >>>> >>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>> To unsubscribe, go to >>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>> and follow the instructions there. >>>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@lists.febo.com >>> To unsubscribe, go to >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>> and follow the instructions there. >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
TP
Tobias Pluess
Fri, Apr 3, 2020 3:20 PM

Hi Bob

knowing that my counter's noise floor is terrible (even though I still
don't understand why) I tried to measure the ADEV and MDEV of my GPSDO
against another GPSDO.

From the graphs, everything below tau=10s is, I would say, rubbish. But I

tend to mistrust these complete results, as I have no means of finding out
whether my reference is so bad or my own GPSDO. The reference is an eBay
GPSDO, and as we all know, these are sometimes of doubtful pedigree.
But still, below the 10s tau, the ADEV and MDEV are so close to the noise
floor that I would say this measurement is useless.

But it still does not explain why my 5335A is so bad.

Tobias
HB9FSX

On Thu, Apr 2, 2020 at 10:17 PM Bob kb8tq kb8tq@n1k.org wrote:

Hi

What you have measured is the noise floor of a 5335 when trying
to use it to measure ADEV. Anything past the numbers on your plot
will be “past” what the 5335 can “see”. Indeed, even when you get
close to those numbers, things may get a bit weird due to the fact
that you are measuring counter “noise” plus device noise.

Bob

On Apr 2, 2020, at 3:13 PM, Tobias Pluess tpluess@ieee.org wrote:

Hello all

in the meantime I figured out most of my problems and my GPSDO is working
now with some very ugly prototype code. Today, I wanted to do some ADEV
measurements.
My plan was to compare the 1PPS generated from my GPSDO to the 1PPS of my
Oscilloquartz STAR4; unfortunately I have nothing else (like Rb or so)
which is perhaps more stable. So I try with the STAR4 and see where I

get.

However, before I did any meaningful measurements, I wanted to see what

the

noise floor of my test equipment is.
Again, unfortunately I have nothing better than a HP 5335A with 1ns
resolution in TIC mode. I measured the noise floor of the TIC as follows:
the 1PPS output of my GPSDO was connected to a resistive power splitter,
and then, one output of the splitter went to channel A of the TIC (START
signal) while the other output from the splitter went first to a long

cable

and then to channel B. With this, I achieved about 16ns of delay.
I then used the TIC together with Timelab and measured the ADEV of this
setup.
As far as I understand, if the delay of the cable stays constant (which

it

does as long as it is not moved and the temperature stays the same), all

I

see in the ADEV plot is the ADEV of my counter itself. Right?

So I let this test run for one hour (collected 3600 samples), and the
result looks terrible. See the attached file. I did the test twice; once

I

used the STAR4 GPSDO as external reference for the counter, and once I

used

its internal reference, which is a HP 10544A oven. As one can see, the

ADEV

at 1sec is between 7e-10 and 8e-10. I don't know yet what numbers I can
expect from my GPSDO, but from datasheets of commercial GPSDOs I saw that
the ADEV shortly after powerup should be in the 1e-11 region. So how does
one measure such low ADEVs?

To me, it appears that the ADEV at 1sec is roughly the counter's
resolution; a bit less due to averaging. If I take averaging over 3600
samples into account, I think I could expect maybe ~1ns/sqrt(3600) =
16.7e-12 as ADEV at 1 second, but we can clearly see that this is not the
case. So there are two interesting questions arising:

a) I think the ADEV is so high because of the quantization error of the
counter. Assume the time interval measured is right at the transition

from,

say, 15ns to 16ns, even the smallest amount of noise will produce some
alternating readings of 15ns and 16ns, which, in turn, results in an ADEV
around 1e-9, right? Further, why is this effect not averaged out with
sqrt(# of samples)?

b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution

does

my counter need? If the above was true, I would expect that a 1ps
resolution (and an even better stability!) was required to measure ADEV

of

1e-12, The fact that the (as far as I know) world's most recent,
rocket-science grade counter (some Keysight stuff) has "only" 20ps of
resolution, but people are still able to measure even 1e-14 shows that my
assumption is wrong. So how are the measurement resolution and the ADEV
related to each other? I plan to build my own TIC based on a TDC7200,

which

would offer some 55ps of resolution, but how low could I go with that?

Best regards
Tobias
HB9FSX

On Fri, Mar 20, 2020 at 5:34 PM Bob Q bobqhome@live.com wrote:

I have seen differences between both UCT and Oscilloquartz 8663 ocxo’s.
The attached plot shows an example. Both boxes use Ublox LEA-6T

receiver,

surveyed in, AD5680 DAC 18 bit DAC, same level shift circuit and same
control circuit. The reference is an LPRO-101. The Oscilloquartz ocxo

was

purchased used. Both UCT ocxo’s (only the better one is shown) were
purchased new and have 100’s of operating hours. I have also seen
differences with constant EFC control voltage. The differences limit

what

performance you can

achieve._______________________________________________

time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

<figure-1.png>_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi Bob knowing that my counter's noise floor is terrible (even though I still don't understand why) I tried to measure the ADEV and MDEV of my GPSDO against another GPSDO. >From the graphs, everything below tau=10s is, I would say, rubbish. But I tend to mistrust these complete results, as I have no means of finding out whether my reference is so bad or my own GPSDO. The reference is an eBay GPSDO, and as we all know, these are sometimes of doubtful pedigree. But still, below the 10s tau, the ADEV and MDEV are so close to the noise floor that I would say this measurement is useless. But it still does not explain why my 5335A is so bad. Tobias HB9FSX On Thu, Apr 2, 2020 at 10:17 PM Bob kb8tq <kb8tq@n1k.org> wrote: > Hi > > What you have measured *is* the noise floor of a 5335 when trying > to use it to measure ADEV. Anything past the numbers on your plot > will be “past” what the 5335 can “see”. Indeed, even when you get > close to those numbers, things may get a bit weird due to the fact > that you are measuring counter “noise” plus device noise. > > Bob > > > On Apr 2, 2020, at 3:13 PM, Tobias Pluess <tpluess@ieee.org> wrote: > > > > Hello all > > > > in the meantime I figured out most of my problems and my GPSDO is working > > now with some very ugly prototype code. Today, I wanted to do some ADEV > > measurements. > > My plan was to compare the 1PPS generated from my GPSDO to the 1PPS of my > > Oscilloquartz STAR4; unfortunately I have nothing else (like Rb or so) > > which is perhaps more stable. So I try with the STAR4 and see where I > get. > > However, before I did any meaningful measurements, I wanted to see what > the > > noise floor of my test equipment is. > > Again, unfortunately I have nothing better than a HP 5335A with 1ns > > resolution in TIC mode. I measured the noise floor of the TIC as follows: > > the 1PPS output of my GPSDO was connected to a resistive power splitter, > > and then, one output of the splitter went to channel A of the TIC (START > > signal) while the other output from the splitter went first to a long > cable > > and then to channel B. With this, I achieved about 16ns of delay. > > I then used the TIC together with Timelab and measured the ADEV of this > > setup. > > As far as I understand, if the delay of the cable stays constant (which > it > > does as long as it is not moved and the temperature stays the same), all > I > > see in the ADEV plot is the ADEV of my counter itself. Right? > > > > So I let this test run for one hour (collected 3600 samples), and the > > result looks terrible. See the attached file. I did the test twice; once > I > > used the STAR4 GPSDO as external reference for the counter, and once I > used > > its internal reference, which is a HP 10544A oven. As one can see, the > ADEV > > at 1sec is between 7e-10 and 8e-10. I don't know yet what numbers I can > > expect from my GPSDO, but from datasheets of commercial GPSDOs I saw that > > the ADEV shortly after powerup should be in the 1e-11 region. So how does > > one measure such low ADEVs? > > > > To me, it appears that the ADEV at 1sec is roughly the counter's > > resolution; a bit less due to averaging. If I take averaging over 3600 > > samples into account, I think I could expect maybe ~1ns/sqrt(3600) = > > 16.7e-12 as ADEV at 1 second, but we can clearly see that this is not the > > case. So there are two interesting questions arising: > > > > a) I think the ADEV is so high because of the quantization error of the > > counter. Assume the time interval measured is right at the transition > from, > > say, 15ns to 16ns, even the smallest amount of noise will produce some > > alternating readings of 15ns and 16ns, which, in turn, results in an ADEV > > around 1e-9, right? Further, why is this effect not averaged out with > > sqrt(# of samples)? > > > > b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution > does > > my counter need? If the above was true, I would expect that a 1ps > > resolution (and an even better stability!) was required to measure ADEV > of > > 1e-12, The fact that the (as far as I know) world's most recent, > > rocket-science grade counter (some Keysight stuff) has "only" 20ps of > > resolution, but people are still able to measure even 1e-14 shows that my > > assumption is wrong. So how are the measurement resolution and the ADEV > > related to each other? I plan to build my own TIC based on a TDC7200, > which > > would offer some 55ps of resolution, but how low could I go with that? > > > > > > Best regards > > Tobias > > HB9FSX > > > > > > > > On Fri, Mar 20, 2020 at 5:34 PM Bob Q <bobqhome@live.com> wrote: > > > >> I have seen differences between both UCT and Oscilloquartz 8663 ocxo’s. > >> The attached plot shows an example. Both boxes use Ublox LEA-6T > receiver, > >> surveyed in, AD5680 DAC 18 bit DAC, same level shift circuit and same > >> control circuit. The reference is an LPRO-101. The Oscilloquartz ocxo > was > >> purchased used. Both UCT ocxo’s (only the better one is shown) were > >> purchased new and have 100’s of operating hours. I have also seen > >> differences with constant EFC control voltage. The differences limit > what > >> performance you can > achieve._______________________________________________ > >> time-nuts mailing list -- time-nuts@lists.febo.com > >> To unsubscribe, go to > >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >> and follow the instructions there. > >> > > <figure-1.png>_______________________________________________ > > time-nuts mailing list -- time-nuts@lists.febo.com > > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >
TP
Tobias Pluess
Fri, Apr 3, 2020 3:25 PM

Hi again Bob,

yes you describe a simple DMTD measurement. But could you tell me what the
difference is between that and comparing the 1PPS pulses?
I mean, I could set the 10811 high in frequency by just 1Hz, and then it
would result in two 1Hz signals which are then compared.
Which is essentially the same as comparing two 1PPS signals, isn't it?
Ok there is a minor difference: since the 1PPS signals are divided down
from 10MHz, their noise is also divided down, which is not the case for the
DMTD.
However, in the end I am comparing signals in the 1Hz to 5Hz or 10Hz
region, and apparently, the 5335A is not suitable for those, at least not
with the desired stability, is it?

Tobias

On Fri, Apr 3, 2020 at 1:45 PM Bob kb8tq kb8tq@n1k.org wrote:

Hi

The quick way to do this is with a single mixer. Take something like an old
10811 and use the coarse tune to set it high in frequency by 5 to 10 Hz.

Then feed it into an RPD-1 mixer and pull out the 5 to 10 Hz audio tone.
That tone is the difference between the 10811 and your device under
test.
If the DUT moves 1 Hz, the audio tone changes by 1 Hz.

If you measured the 10 MHz on the DUT, that 1 Hz would be a very small
shift
( 0.1 ppm ). At 10 Hz it’s a 10% change. You have “amplified” the change
in frequency by the ratio of 10 MHz to 10 Hz ( so a million X increase ).

IF you could tack that on to the ADEV plot of your 5335 ( no, it’s not
that
simple) your 7x10^-10 at 1 second would become more 7x10^-16 at 1
second.

The reason its not quite that simple is that the input circuit on the
counter
really does not handle a 10 Hz audio tone as well as it handles a 10 MHz
RF signal. Instead of getting 9 digits a second, you probably will get
three
good digits a second and another 6 digits of noise.

The good news is that an op amp used as a preamp ( to get you up to maybe
32 V p-p rather than a volt or so) and another op amp or three as limiters
will
get you up around 6 or 7 good digits. Toss in a cap or two as a high pass
and low pass filter ( DC offsets can be a problem ….) and you have a
working
device that gets into the parts in 10^-13 with your 5335.

It all can be done with point to point wiring. No need for a PCB layout.
Be
careful that the +/- 18V supplies to the op amp both go on and off at
the
same time ….

Bob

On Apr 3, 2020, at 5:13 AM, Tobias Pluess tpluess@ieee.org wrote:

hi John

yes I know the DMTD method, and indeed I am planing to build my own DMTD
system, something similar to the "Small DMTD system" published by Riley (
https://www.wriley.com/A Small DMTD System.pdf).
However I am unsure whether that will help much in this case, because all
what the DMTD does is to mix the 10MHz signals down to some 1Hz Signal or
so which can be measured more easily, and I already have 1Hz signals (the
1PPS) which I am comparing.
Or do you suggest to use the DMTD and use a higher frequency at its
outputs, say 10Hz or so, and then average for 10 samples  to increase the
resolution?

Thanks
Tobias
HB9FSX

On Fri, Apr 3, 2020 at 12:53 AM John Miles john@miles.io wrote:

b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution

does

my counter need? If the above was true, I would expect that a 1ps
resolution (and an even better stability!) was required to measure ADEV

of

1e-12, The fact that the (as far as I know) world's most recent,
rocket-science grade counter (some Keysight stuff) has "only" 20ps of
resolution, but people are still able to measure even 1e-14 shows that

my

assumption is wrong. So how are the measurement resolution and the ADEV
related to each other? I plan to build my own TIC based on a TDC7200,

which

would offer some 55ps of resolution, but how low could I go with that?

That sounds like a simple question but it's not.  There are a few
different approaches to look into:

  1. Use averaging with your existing counter.  Some counters can yield
    readings in the 1E-12 region at t=1s even though their single-shot

jitter

is much worse than that.  They do this by averaging  hundreds or

thousands

of samples for each reading they report.  Whether (and when) this is
acceptable is a complex topic in itself, too much so to explain quickly.
Search for information on the effects of averaging and dead time on

Allan

deviation to find the entrance to this fork of the rabbit hole.

  1. Search for the term 'DMTD' and read about that.

  2. Search for 'direct digital phase measurement' and read about that.

  3. Search for 'tight PLL' and read about that.

Basically, while some counters can perform averaging on a post-detection
basis, that's like using the tone control on a radio to reduce static

and

QRM.  It works, sort of, but it's too late in the signal chain at that
point to do the job right.  You really want to limit the bandwidth

before

the signal is captured, but since that's almost never practical at RF,

the

next best thing to do is limit the bandwidth before the signal is
"demodulated" (i.e., counted.)

Hence items 2, 3, and 4 above.  They either limit the measurement
bandwidth prior to detection, lower the frequency itself to keep the
counter's inherent jitter from dominating the measurement, or both.

You'll

have to use one of these methods, or another technique along the same
lines, if you want to measure the short-term stability of a good

oscillator

or GPSDO.

-- john, KE5FX


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi again Bob, yes you describe a simple DMTD measurement. But could you tell me what the difference is between that and comparing the 1PPS pulses? I mean, I could set the 10811 high in frequency by just 1Hz, and then it would result in two 1Hz signals which are then compared. Which is essentially the same as comparing two 1PPS signals, isn't it? Ok there is a minor difference: since the 1PPS signals are divided down from 10MHz, their noise is also divided down, which is not the case for the DMTD. However, in the end I am comparing signals in the 1Hz to 5Hz or 10Hz region, and apparently, the 5335A is not suitable for those, at least not with the desired stability, is it? Tobias On Fri, Apr 3, 2020 at 1:45 PM Bob kb8tq <kb8tq@n1k.org> wrote: > Hi > > The quick way to do this is with a single mixer. Take something like an old > 10811 and use the coarse tune to set it high in frequency by 5 to 10 Hz. > > Then feed it into an RPD-1 mixer and pull out the 5 to 10 Hz audio tone. > That tone is the *difference* between the 10811 and your device under > test. > If the DUT moves 1 Hz, the audio tone changes by 1 Hz. > > If you measured the 10 MHz on the DUT, that 1 Hz would be a very small > shift > ( 0.1 ppm ). At 10 Hz it’s a 10% change. You have “amplified” the change > in frequency by the ratio of 10 MHz to 10 Hz ( so a million X increase ). > > *IF* you could tack that on to the ADEV plot of your 5335 ( no, it’s not > that > simple) your 7x10^-10 at 1 second would become more 7x10^-16 at 1 > second. > > The reason its not quite that simple is that the input circuit on the > counter > really does not handle a 10 Hz audio tone as well as it handles a 10 MHz > RF signal. Instead of getting 9 digits a second, you probably will get > three > *good* digits a second and another 6 digits of noise. > > The good news is that an op amp used as a preamp ( to get you up to maybe > 32 V p-p rather than a volt or so) and another op amp or three as limiters > will > get you up around 6 or 7 good digits. Toss in a cap or two as a high pass > and low pass filter ( DC offsets can be a problem ….) and you have a > working > device that gets into the parts in 10^-13 with your 5335. > > It all can be done with point to point wiring. No need for a PCB layout. > Be > careful that the +/- 18V supplies to the op amp *both* go on and off at > the > same time …. > > Bob > > > On Apr 3, 2020, at 5:13 AM, Tobias Pluess <tpluess@ieee.org> wrote: > > > > hi John > > > > yes I know the DMTD method, and indeed I am planing to build my own DMTD > > system, something similar to the "Small DMTD system" published by Riley ( > > https://www.wriley.com/A Small DMTD System.pdf). > > However I am unsure whether that will help much in this case, because all > > what the DMTD does is to mix the 10MHz signals down to some 1Hz Signal or > > so which can be measured more easily, and I already have 1Hz signals (the > > 1PPS) which I am comparing. > > Or do you suggest to use the DMTD and use a higher frequency at its > > outputs, say 10Hz or so, and then average for 10 samples to increase the > > resolution? > > > > Thanks > > Tobias > > HB9FSX > > > > > > On Fri, Apr 3, 2020 at 12:53 AM John Miles <john@miles.io> wrote: > > > >>> b) if I want to measure 1e-11 or even 1e-12 at 1sec - what resolution > >> does > >>> my counter need? If the above was true, I would expect that a 1ps > >>> resolution (and an even better stability!) was required to measure ADEV > >> of > >>> 1e-12, The fact that the (as far as I know) world's most recent, > >>> rocket-science grade counter (some Keysight stuff) has "only" 20ps of > >>> resolution, but people are still able to measure even 1e-14 shows that > my > >>> assumption is wrong. So how are the measurement resolution and the ADEV > >>> related to each other? I plan to build my own TIC based on a TDC7200, > >> which > >>> would offer some 55ps of resolution, but how low could I go with that? > >> > >> That sounds like a simple question but it's not. There are a few > >> different approaches to look into: > >> > >> 1) Use averaging with your existing counter. Some counters can yield > >> readings in the 1E-12 region at t=1s even though their single-shot > jitter > >> is much worse than that. They do this by averaging hundreds or > thousands > >> of samples for each reading they report. Whether (and when) this is > >> acceptable is a complex topic in itself, too much so to explain quickly. > >> Search for information on the effects of averaging and dead time on > Allan > >> deviation to find the entrance to this fork of the rabbit hole. > >> > >> 2) Search for the term 'DMTD' and read about that. > >> > >> 3) Search for 'direct digital phase measurement' and read about that. > >> > >> 4) Search for 'tight PLL' and read about that. > >> > >> Basically, while some counters can perform averaging on a post-detection > >> basis, that's like using the tone control on a radio to reduce static > and > >> QRM. It works, sort of, but it's too late in the signal chain at that > >> point to do the job right. You really want to limit the bandwidth > before > >> the signal is captured, but since that's almost never practical at RF, > the > >> next best thing to do is limit the bandwidth before the signal is > >> "demodulated" (i.e., counted.) > >> > >> Hence items 2, 3, and 4 above. They either limit the measurement > >> bandwidth prior to detection, lower the frequency itself to keep the > >> counter's inherent jitter from dominating the measurement, or both. > You'll > >> have to use one of these methods, or another technique along the same > >> lines, if you want to measure the short-term stability of a good > oscillator > >> or GPSDO. > >> > >> -- john, KE5FX > >> > >> > >> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@lists.febo.com > >> To unsubscribe, go to > >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >> and follow the instructions there. > >> > > _______________________________________________ > > time-nuts mailing list -- time-nuts@lists.febo.com > > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >