Can any list member please tell me the "accuracy" that can be expected from
a typical GPSDO
over, say, a 10 second interval? I have several measuring instruments
connected to my Trimbal GPSDO, and would like to know what to expect. At
the moment I am guessing about 1 to 2 parts in 10^12.
Thankyou,................................................................................Donald
Brett Collie
https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail
Virus-free.
www.avast.com
https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail
<#DAB4FAD8-2DD7-40BB-A1B8-4E2AA1F9FDF2>
Hi
Depending on your definition of “accuracy”, and the actual device, the answer is likely
in the 1 ppb to 10 ppt range based on a 10 second measurement.
What’s going on?
First up you have the basic accuracy of GPS. Over many days of averaging (with a fairly
fancy averaging device) that can be quite good ( << 1 ppt). Most GPSDO’s run averaging
in the (maybe) 100 second range. That gets you into a whole different arena. ( = long term
accuracy is not the issue).
Next, you have noise. GPS at 1 second likely has several ppb of noise (as measured by ADEV).
At 10 seconds you should be below one ppb, but maybe not by a lot.
Your local reference also has noise as does the loop. Is the loop + reference 1 ppb or 1 ppt at
one second? (again measuring with ADEV). It depends a lot on just what you have. A good guess
is 10 ppt. That may also apply at 10 or even at 30 seconds.
Now comes the question of your “accuracy” definition. ADEV is a “one sigma, delta between readings”
measure. If you are after absolute frequency, this is not what ADEV is measuring. If you are after a
six sigma “frequency accuracy” number, you might well need to multiply the ADEV by 10 or 100X.
Often GPSDO’s that are intended as frequency references are spec’d out as “better than X 99% of
the time ( typical ).” You have a sort of three sigma, with the qualifier that it is typical. ( So not all the
time … hmmm …..).
So what are you trying to do? What sort of “accuracy” do you need? What are you looking for?
Is ADEV an adequate measure?
Bob
On Sep 9, 2020, at 7:04 PM, donald collie donaldbcollie@gmail.com wrote:
Can any list member please tell me the "accuracy" that can be expected from
a typical GPSDO
over, say, a 10 second interval? I have several measuring instruments
connected to my Trimbal GPSDO, and would like to know what to expect. At
the moment I am guessing about 1 to 2 parts in 10^12.
Thankyou,................................................................................Donald
Brett Collie
https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail
Virus-free.
www.avast.com
https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail
<#DAB4FAD8-2DD7-40BB-A1B8-4E2AA1F9FDF2>
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Not sure what you mean by "accuracy over a 10 sec interval". If you mean
(in)correctness of frequency averaged over a given 10 sec interval, it's
more
like a few parts in 10^9. But if you mean stability of frequency without
regard
for its correctness, you might see your 2 parts in 10^12 in at least some
intervals,
although that seems a bit tight to me.
The numbers I mention above are for a GPSDO using an OCXO with a
disciplining
time constant of just a few seconds, which is kind of a worst case
scenario. However,
if your VCO is stable enough to be reigned in with a time constant of
hundreds of
seconds, you can do much better for stability and at least moderately
better for
absolute frequency accuracy.
If your primary interest is fixed in that 10 sec interval, a GPSDO with a
good OCXO is
likely to be the best way to go. OTOH, if your interest evolves into
appreciably longer time
intervals, you'd likely do better with a GPS disciplined Rb oscillator,
where disciplining
time constants of several hours (or even longer) are feasible. However,
many of the
usual Rb standards tend to have a hump in their phase and frequency
variations around
time intervals of a few seconds.
I hope this helps.
Dana
On Wed, Sep 9, 2020 at 6:40 PM donald collie donaldbcollie@gmail.com
wrote:
Can any list member please tell me the "accuracy" that can be expected from
a typical GPSDO
over, say, a 10 second interval? I have several measuring instruments
connected to my Trimbal GPSDO, and would like to know what to expect. At
the moment I am guessing about 1 to 2 parts in 10^12.
Thankyou,................................................................................Donald
Brett Collie
Virus-free.
www.avast.com
<
https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail
<#DAB4FAD8-2DD7-40BB-A1B8-4E2AA1F9FDF2>
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Hi Donald --
Over a 10 second interval, you're really looking at the short term
stability of the oscillator in the GPSDO, which could vary be several
orders of magnitude. The GPS component will set the nominal frequency,
but the noise around nominal will be that of the OCXO.
A quite good GPSDO, like a Z3801A or Trimble Thunderbolt, can show short
term stability in the 12s, but not all will be that good. An
inexpensive unit like the Bodnar might be more like 1 PPB.
On 9/9/20 7:04 PM, donald collie wrote:
Can any list member please tell me the "accuracy" that can be expected from
a typical GPSDO
over, say, a 10 second interval? I have several measuring instruments
connected to my Trimbal GPSDO, and would like to know what to expect. At
the moment I am guessing about 1 to 2 parts in 10^12.
Thankyou,................................................................................Donald
Brett Collie
https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail
Virus-free.
www.avast.com
https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail
<#DAB4FAD8-2DD7-40BB-A1B8-4E2AA1F9FDF2>
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Hi
Maybe a good idea to “back up” a bit here:
The most commonly plotted data for the performance of a GPSDO is ADEV.
Very simply put, to do ADEV you take a series of readings at a specific time
spacing ( called tau ). The delta frequency from one reading to the next is then
computed. You take the standard deviation of that “delta” information.
What you are looking at is the “good guess” at what the frequency will be in the
next time slot, based on what it is in this time slot. This may or may not be what
your system / measurement instruments are looking as a. spec.
The big reason ADEV exists is that it is convergent. As you take more data, the
results don’t move all over the place, they converge to a single value. Measure today,
then measure tomorrow, you get pretty much the same number.
You might want to know what the maximum frequency error compared to “absolutely
correct” is in your 10 second time period. This is a measure that is non-convergent. The
longer you collect the data, the bigger the number gets. Measure for an hour and you
get a different number compared to measuring for a day. Measure today / measure
tomorrow and you may get dramatically different results.
This is not to say that nobody ever can know what the frequency is. Only that “max” is
not a good limit for this sort of random fluctuation. Again, this is what drove the guys
at NIST to come up with ADEV back in the 1960’s. It’s what keeps us using it as a
means of comparison today.
Bob
On Sep 9, 2020, at 7:04 PM, donald collie donaldbcollie@gmail.com wrote:
Can any list member please tell me the "accuracy" that can be expected from
a typical GPSDO
over, say, a 10 second interval? I have several measuring instruments
connected to my Trimbal GPSDO, and would like to know what to expect. At
the moment I am guessing about 1 to 2 parts in 10^12.
Thankyou,................................................................................Donald
Brett Collie
https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail
Virus-free.
www.avast.com
https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail
<#DAB4FAD8-2DD7-40BB-A1B8-4E2AA1F9FDF2>
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Newbie here looking for comments from people with more experience with
thunderbolt E's to comment on what I am seeing.
Unit is jumping and ramping the DAC one direction then goes stable for a
bit only to start jumping and ramping the other way.
Oscillator heading south or more likely the electronics package?
Was running OK, shutdown and put away for months and then restarted and
has been running like this for months and not settling down.
Thanks
Kevin
Hi
A couple things to check:
Is your power supply doing odd things? It’s the first thing I’d suspect in this case.
Does the OCXO “behave” if you turn off the GPS ( = disconnect the antenna ) and watch
it with a counter?
While the antenna is disconnected from the TBolt, hook it to something else and see how it’s doing?
What does. LH show for sats and all the other stuff as this is going on ?
Bob
On Sep 10, 2020, at 2:40 PM, Kevin Schuchmann kschuchm@gmail.com wrote:
Newbie here looking for comments from people with more experience with thunderbolt E's to comment on what I am seeing.
Unit is jumping and ramping the DAC one direction then goes stable for a bit only to start jumping and ramping the other way.
Oscillator heading south or more likely the electronics package?
Was running OK, shutdown and put away for months and then restarted and has been running like this for months and not settling down.
Thanks
Kevin
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Guess my image didn't make it, I will add it as an attachment this time.
Swapped power supplies no difference.
Antenna is on a splitter shared with a non E thunderbolt and it is
working fine along with a NTP100 and Tymserve 2100 with no complaints
from them.
I will try disconnecting the antenna and see what happens.
Thanks for the suggestions.
Kevin
On 9/10/2020 2:41 PM, Bob kb8tq wrote:
Hi
A couple things to check:
Is your power supply doing odd things? It’s the first thing I’d suspect in this case.
Does the OCXO “behave” if you turn off the GPS ( = disconnect the antenna ) and watch
it with a counter?
While the antenna is disconnected from the TBolt, hook it to something else and see how it’s doing?
What does. LH show for sats and all the other stuff as this is going on ?
Bob
On Sep 10, 2020, at 2:40 PM, Kevin Schuchmann kschuchm@gmail.com wrote:
Newbie here looking for comments from people with more experience with thunderbolt E's to comment on what I am seeing.
Unit is jumping and ramping the DAC one direction then goes stable for a bit only to start jumping and ramping the other way.
Oscillator heading south or more likely the electronics package?
Was running OK, shutdown and put away for months and then restarted and has been running like this for months and not settling down.
Thanks
Kevin
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Thankyou to all who responded. It looks as if 2 parts in 10^12 is about
what can be expected - certainly much better than the 1 part in 10^9
available from a typical double ovened quartz crystal oscillator barefoot.
Cheers!...............................................................................................................................Donald
Brett Collie
https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail
Virus-free.
www.avast.com
https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail
<#DAB4FAD8-2DD7-40BB-A1B8-4E2AA1F9FDF2>
On Fri, Sep 11, 2020 at 1:15 AM Bob kb8tq kb8tq@n1k.org wrote:
Hi
Maybe a good idea to “back up” a bit here:
The most commonly plotted data for the performance of a GPSDO is ADEV.
Very simply put, to do ADEV you take a series of readings at a specific
time
spacing ( called tau ). The delta frequency from one reading to the next
is then
computed. You take the standard deviation of that “delta” information.
What you are looking at is the “good guess” at what the frequency will be
in the
next time slot, based on what it is in this time slot. This may or may not
be what
your system / measurement instruments are looking as a. spec.
The big reason ADEV exists is that it is convergent. As you take more
data, the
results don’t move all over the place, they converge to a single value.
Measure today,
then measure tomorrow, you get pretty much the same number.
You might want to know what the maximum frequency error compared to
“absolutely
correct” is in your 10 second time period. This is a measure that is
non-convergent. The
longer you collect the data, the bigger the number gets. Measure for an
hour and you
get a different number compared to measuring for a day. Measure today /
measure
tomorrow and you may get dramatically different results.
This is not to say that nobody ever can know what the frequency is. Only
that “max” is
not a good limit for this sort of random fluctuation. Again, this is what
drove the guys
at NIST to come up with ADEV back in the 1960’s. It’s what keeps us using
it as a
means of comparison today.
Bob
On Sep 9, 2020, at 7:04 PM, donald collie donaldbcollie@gmail.com
wrote:
Can any list member please tell me the "accuracy" that can be expected
from
a typical GPSDO
over, say, a 10 second interval? I have several measuring instruments
connected to my Trimbal GPSDO, and would like to know what to expect. At
the moment I am guessing about 1 to 2 parts in 10^12.
Thankyou,................................................................................Donald
Brett Collie
<
Virus-free.
www.avast.com
<
<#DAB4FAD8-2DD7-40BB-A1B8-4E2AA1F9FDF2>
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
On Freitag, 11. September 2020 01:08:09 CEST Kevin Schuchmann wrote:
Guess my image didn't make it, I will add it as an attachment this time.
The temperature curve seems to show some correlation to what is happening with
the DAC. Seems that the DAC and OSC jumps are mostly during periods of some
thermal perturbation.
Regards,
Matthias
Hi
Well ….
A reasonably good DOCXO will come in around 2x10^-10 over 0 to 70C. Over a
typical 4 C “lab wander” you should see < 1x10^-11. After being on power for a month
or three, aging should be below 1x10^-11 per day.
ADEV wise, it should do better than the GPSDO …..parts in 10^-13 are not unusual at
10 seconds on a typical design.
Bob
On Sep 11, 2020, at 3:35 AM, donald collie donaldbcollie@gmail.com wrote:
Thankyou to all who responded. It looks as if 2 parts in 10^12 is about
what can be expected - certainly much better than the 1 part in 10^9
available from a typical double ovened quartz crystal oscillator barefoot.
Cheers!...............................................................................................................................Donald
Brett Collie
https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail
Virus-free.
www.avast.com
https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail
<#DAB4FAD8-2DD7-40BB-A1B8-4E2AA1F9FDF2>
On Fri, Sep 11, 2020 at 1:15 AM Bob kb8tq kb8tq@n1k.org wrote:
Hi
Maybe a good idea to “back up” a bit here:
The most commonly plotted data for the performance of a GPSDO is ADEV.
Very simply put, to do ADEV you take a series of readings at a specific
time
spacing ( called tau ). The delta frequency from one reading to the next
is then
computed. You take the standard deviation of that “delta” information.
What you are looking at is the “good guess” at what the frequency will be
in the
next time slot, based on what it is in this time slot. This may or may not
be what
your system / measurement instruments are looking as a. spec.
The big reason ADEV exists is that it is convergent. As you take more
data, the
results don’t move all over the place, they converge to a single value.
Measure today,
then measure tomorrow, you get pretty much the same number.
You might want to know what the maximum frequency error compared to
“absolutely
correct” is in your 10 second time period. This is a measure that is
non-convergent. The
longer you collect the data, the bigger the number gets. Measure for an
hour and you
get a different number compared to measuring for a day. Measure today /
measure
tomorrow and you may get dramatically different results.
This is not to say that nobody ever can know what the frequency is. Only
that “max” is
not a good limit for this sort of random fluctuation. Again, this is what
drove the guys
at NIST to come up with ADEV back in the 1960’s. It’s what keeps us using
it as a
means of comparison today.
Bob
On Sep 9, 2020, at 7:04 PM, donald collie donaldbcollie@gmail.com
wrote:
Can any list member please tell me the "accuracy" that can be expected
from
a typical GPSDO
over, say, a 10 second interval? I have several measuring instruments
connected to my Trimbal GPSDO, and would like to know what to expect. At
the moment I am guessing about 1 to 2 parts in 10^12.
Thankyou,................................................................................Donald
Brett Collie
<
Virus-free.
www.avast.com
<
<#DAB4FAD8-2DD7-40BB-A1B8-4E2AA1F9FDF2>
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Matthias,
You are correct, having spent hours looking at all the stable areas
and cooling and heating the gpsdo I find that when it reports that it is
116.25 F then it is stable.
So now I guess I need to figure out why it is so picky. I will measure
the current draw from a cold start and see if I see the oven warming up
and then stabilizing and then heat and cool it and see how it reacts,
and also look at the electronics and see if an area is overly sensitive
to temperature.
Thanks
Kevin
On 9/11/2020 2:26 AM, Matthias Welwarsky wrote:
On Freitag, 11. September 2020 01:08:09 CEST Kevin Schuchmann wrote:
Guess my image didn't make it, I will add it as an attachment this time.
The temperature curve seems to show some correlation to what is happening with
the DAC. Seems that the DAC and OSC jumps are mostly during periods of some
thermal perturbation.
Regards,
Matthias
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
I've been looking for a thunderbolt for a while, mostly because they're
such a benchmark within this group. I missed the period when they were more
commonly available and have mostly found ones that were a lot more
expensive, or had a large delivery cost due to being in Australia.
Some cheaper examples turned up on UK ebay. These aren't the classic
time-nuts ones but presumably look fairly similar to Lady Heather.
https://www.ebay.co.uk/itm/154078750287
I haven't been able to find a comparison of the various thunderbolt models.
Is there a description somewhere ? I've found a few hints about this one -
I think it's a single 48V supply and possibly a lower-quality OCXO.
On Sat, Sep 12, 2020 at 5:14 PM Kevin Schuchmann kschuchm@gmail.com wrote:
Matthias,
You are correct, having spent hours looking at all the stable areas
and cooling and heating the gpsdo I find that when it reports that it is
116.25 F then it is stable.
So now I guess I need to figure out why it is so picky. I will measure
the current draw from a cold start and see if I see the oven warming up
and then stabilizing and then heat and cool it and see how it reacts,
and also look at the electronics and see if an area is overly sensitive
to temperature.
Thanks
Kevin
On 9/11/2020 2:26 AM, Matthias Welwarsky wrote:
On Freitag, 11. September 2020 01:08:09 CEST Kevin Schuchmann wrote:
Guess my image didn't make it, I will add it as an attachment this time.
The temperature curve seems to show some correlation to what is
happening with
the DAC. Seems that the DAC and OSC jumps are mostly during periods of
some
thermal perturbation.
Regards,
Matthias
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Hi
The unit your link points to is the “classic” TBolt that showed up in volume a bit
over a decade ago. That one is packaged with a (noisy) switching power supply.
There is no easy way to tell what era it is from. Indeed if it is from the 1990’s the
OCXO will not be quite as good as what’s in one from > 2002. Since the unit normally
operates locked, that’s not a real big deal.
Since Trimble used “Thunderbolt” for a whole line of parts and was not very good
at labeling devices, there is no easy way to refer to this or that specific device. That
makes coming up with a list of units a bit hard. For example - what’s the model number
of the part in that eBay listing ? ( No I don’t know either …. :) ).
Bob
On Sep 12, 2020, at 12:28 PM, Adrian Godwin artgodwin@gmail.com wrote:
I've been looking for a thunderbolt for a while, mostly because they're
such a benchmark within this group. I missed the period when they were more
commonly available and have mostly found ones that were a lot more
expensive, or had a large delivery cost due to being in Australia.
Some cheaper examples turned up on UK ebay. These aren't the classic
time-nuts ones but presumably look fairly similar to Lady Heather.
https://www.ebay.co.uk/itm/154078750287
I haven't been able to find a comparison of the various thunderbolt models.
Is there a description somewhere ? I've found a few hints about this one -
I think it's a single 48V supply and possibly a lower-quality OCXO.
On Sat, Sep 12, 2020 at 5:14 PM Kevin Schuchmann kschuchm@gmail.com wrote:
Matthias,
You are correct, having spent hours looking at all the stable areas
and cooling and heating the gpsdo I find that when it reports that it is
116.25 F then it is stable.
So now I guess I need to figure out why it is so picky. I will measure
the current draw from a cold start and see if I see the oven warming up
and then stabilizing and then heat and cool it and see how it reacts,
and also look at the electronics and see if an area is overly sensitive
to temperature.
Thanks
Kevin
On 9/11/2020 2:26 AM, Matthias Welwarsky wrote:
On Freitag, 11. September 2020 01:08:09 CEST Kevin Schuchmann wrote:
Guess my image didn't make it, I will add it as an attachment this time.
The temperature curve seems to show some correlation to what is
happening with
the DAC. Seems that the DAC and OSC jumps are mostly during periods of
some
thermal perturbation.
Regards,
Matthias
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Kevin,
how long has it been in storage? I expect the GPSDO doing quite clever things
to compensate for temperature effects. There may be some parameters stored on
the device and the characteristics of the OCXO may have changed after being
turned off for an extended period.
Regards,
Matthias
On Samstag, 12. September 2020 18:13:35 CEST Kevin Schuchmann wrote:
Matthias,
You are correct, having spent hours looking at all the stable areas
and cooling and heating the gpsdo I find that when it reports that it is
116.25 F then it is stable.
So now I guess I need to figure out why it is so picky. I will measure
the current draw from a cold start and see if I see the oven warming up
and then stabilizing and then heat and cool it and see how it reacts,
and also look at the electronics and see if an area is overly sensitive
to temperature.
Thanks
Kevin
On 9/11/2020 2:26 AM, Matthias Welwarsky wrote:
On Freitag, 11. September 2020 01:08:09 CEST Kevin Schuchmann wrote:
Guess my image didn't make it, I will add it as an attachment this time.
The temperature curve seems to show some correlation to what is happening
with the DAC. Seems that the DAC and OSC jumps are mostly during periods
of some thermal perturbation.
Regards,
Matthias
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and
follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow
the instructions there.
Kevin,
Can you post your Lady Heather config file and command line options that
created that image ?
George
On 9/10/2020 19:08, Kevin Schuchmann wrote:
Guess my image didn't make it, I will add it as an attachment this time.
Swapped power supplies no difference.
Antenna is on a splitter shared with a non E thunderbolt and it is
working fine along with a NTP100 and Tymserve 2100 with no complaints
from them.
I will try disconnecting the antenna and see what happens.
Thanks for the suggestions.
Kevin
On 9/10/2020 2:41 PM, Bob kb8tq wrote:
Hi
A couple things to check:
Is your power supply doing odd things? It’s the first thing I’d
suspect in this case.
Does the OCXO “behave” if you turn off the GPS ( = disconnect the
antenna ) and watch
it with a counter?
While the antenna is disconnected from the TBolt, hook it to
something else and see how it’s doing?
What does. LH show for sats and all the other stuff as this is
going on ?
Bob
On Sep 10, 2020, at 2:40 PM, Kevin Schuchmann kschuchm@gmail.com
wrote:
Newbie here looking for comments from people with more experience
with thunderbolt E's to comment on what I am seeing.
Unit is jumping and ramping the DAC one direction then goes stable
for a bit only to start jumping and ramping the other way.
Oscillator heading south or more likely the electronics package?
Was running OK, shutdown and put away for months and then restarted
and has been running like this for months and not settling down.
Thanks
Kevin
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Thanks!
The one I've seen pictured most often comes in a (I think) slightly lower
gold-coloured (alocromed) case.
e.g. http://www.sydneystormcity.com/TrimbleThunderboltGPS.jpg
Is that any different ?
I did order the one from my link so will report on any markings. The vendor
mentioned part 48050-61 but as you say, that seems to be common to all of
them.
On Sat, Sep 12, 2020 at 10:14 PM Bob kb8tq kb8tq@n1k.org wrote:
Hi
The unit your link points to is the “classic” TBolt that showed up in
volume a bit
over a decade ago. That one is packaged with a (noisy) switching power
supply.
There is no easy way to tell what era it is from. Indeed if it is from the
1990’s the
OCXO will not be quite as good as what’s in one from > 2002. Since the
unit normally
operates locked, that’s not a real big deal.
Since Trimble used “Thunderbolt” for a whole line of parts and was not
very good
at labeling devices, there is no easy way to refer to this or that
specific device. That
makes coming up with a list of units a bit hard. For example - what’s the
model number
of the part in that eBay listing ? ( No I don’t know either …. :) ).
Bob
On Sep 12, 2020, at 12:28 PM, Adrian Godwin artgodwin@gmail.com wrote:
I've been looking for a thunderbolt for a while, mostly because they're
such a benchmark within this group. I missed the period when they were
more
commonly available and have mostly found ones that were a lot more
expensive, or had a large delivery cost due to being in Australia.
Some cheaper examples turned up on UK ebay. These aren't the classic
time-nuts ones but presumably look fairly similar to Lady Heather.
https://www.ebay.co.uk/itm/154078750287
I haven't been able to find a comparison of the various thunderbolt
models.
Is there a description somewhere ? I've found a few hints about this one
I think it's a single 48V supply and possibly a lower-quality OCXO.
On Sat, Sep 12, 2020 at 5:14 PM Kevin Schuchmann kschuchm@gmail.com
wrote:
Matthias,
You are correct, having spent hours looking at all the stable areas
and cooling and heating the gpsdo I find that when it reports that it is
116.25 F then it is stable.
So now I guess I need to figure out why it is so picky. I will measure
the current draw from a cold start and see if I see the oven warming up
and then stabilizing and then heat and cool it and see how it reacts,
and also look at the electronics and see if an area is overly sensitive
to temperature.
Thanks
Kevin
On 9/11/2020 2:26 AM, Matthias Welwarsky wrote:
On Freitag, 11. September 2020 01:08:09 CEST Kevin Schuchmann wrote:
Guess my image didn't make it, I will add it as an attachment this
time.
The temperature curve seems to show some correlation to what is
happening with
the DAC. Seems that the DAC and OSC jumps are mostly during periods of
some
thermal perturbation.
Regards,
Matthias
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Hi
As far as anybody knows (which is a big qualifier indeed ….), the TBolt family only
does it’s “temperature compensation” stuff when it’s in holdover. While it’s locked, there
appears to be no compensation being done. If it’s doing anything while locked, it’s
learning what the TC is, so it can use that information while in holdover.
Bob
On Sep 12, 2020, at 3:52 PM, Matthias Welwarsky time-nuts@welwarsky.de wrote:
Kevin,
how long has it been in storage? I expect the GPSDO doing quite clever things
to compensate for temperature effects. There may be some parameters stored on
the device and the characteristics of the OCXO may have changed after being
turned off for an extended period.
Regards,
Matthias
On Samstag, 12. September 2020 18:13:35 CEST Kevin Schuchmann wrote:
Matthias,
You are correct, having spent hours looking at all the stable areas
and cooling and heating the gpsdo I find that when it reports that it is
116.25 F then it is stable.
So now I guess I need to figure out why it is so picky. I will measure
the current draw from a cold start and see if I see the oven warming up
and then stabilizing and then heat and cool it and see how it reacts,
and also look at the electronics and see if an area is overly sensitive
to temperature.
Thanks
Kevin
On 9/11/2020 2:26 AM, Matthias Welwarsky wrote:
On Freitag, 11. September 2020 01:08:09 CEST Kevin Schuchmann wrote:
Guess my image didn't make it, I will add it as an attachment this time.
The temperature curve seems to show some correlation to what is happening
with the DAC. Seems that the DAC and OSC jumps are mostly during periods
of some thermal perturbation.
Regards,
Matthias
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and
follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow
the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Matthias,
It was put away for about 3 months, then powered up and ignored for
about a month as I figured it would have to settle in again.
I then started monitoring it with LH and saw the abrupt jumps that
would not go away.
I pulled the cover off today and the OCXO behaves as expected, cool
it and the current draw goes up, warm it and current drops.
It is stable as far as I can tell with the equipment I have.
What I did find is the oscillator for the GPS ingest is incredibly
sensitive to temp changes, just blowing a little air through a straw at
it can cause the unit to no longer find any usable satellites. I then
started paying more attention to the Doppler values and realized this is
where the abrupt shifts are coming from, they should be a smooth
increase or decrease depending on if the sat is approaching or receding.
I totally missed that they are glitching, if I can hit the sweet spot
temp wise the GPS Doppler becomes smooth with no quick jumps and the
chart reflects this with a stable oscillator plot.
So it looks like my problem is the GPS ingest section, looking at it's
oscillator it measures 12.5 MHz but I have not been able to find any
specs so far based on the numbers on its case.
Thanks
Kevin
On 9/12/2020 12:52 PM, Matthias Welwarsky wrote:
Kevin,
how long has it been in storage? I expect the GPSDO doing quite clever things
to compensate for temperature effects. There may be some parameters stored on
the device and the characteristics of the OCXO may have changed after being
turned off for an extended period.
Regards,
Matthias
On Samstag, 12. September 2020 18:13:35 CEST Kevin Schuchmann wrote:
Matthias,
You are correct, having spent hours looking at all the stable areas
and cooling and heating the gpsdo I find that when it reports that it is
116.25 F then it is stable.
So now I guess I need to figure out why it is so picky. I will measure
the current draw from a cold start and see if I see the oven warming up
and then stabilizing and then heat and cool it and see how it reacts,
and also look at the electronics and see if an area is overly sensitive
to temperature.
Thanks
Kevin
Hi
The unit you most commonly see is a version done for E911 cell tower upgrades.
It was an intermediate step to add that functionality before it could be done internal
to the cell base station. Once the “real” solution came along, the E911 boxes became
scrap. They went off to China to be salvaged.
The one in the listing is simply the board you normally see mated up with a power supply.
Trimble shipped them out mainly as evaluation devices. The bulk of what was produced
went into the E911 application.
There is a bunch of information on all this back in the archives 10 to 15 years ago. Searching
those archives is a bit of a challenge with all the changes in the list over the years.
Functionally, there is no difference between the one you are looking at and the more common
one. Both came in a variety of firmware versions. Both have a range of OCXO’s in them
(99% all from the same source, but changing designs over the years …). Without digging
into the exact board, there is no easy way to see what’s what. The good news is that they
all pretty much work fine.
Bob
On Sep 12, 2020, at 5:43 PM, Adrian Godwin artgodwin@gmail.com wrote:
Thanks!
The one I've seen pictured most often comes in a (I think) slightly lower
gold-coloured (alocromed) case.
e.g. http://www.sydneystormcity.com/TrimbleThunderboltGPS.jpg
Is that any different ?
I did order the one from my link so will report on any markings. The vendor
mentioned part 48050-61 but as you say, that seems to be common to all of
them.
On Sat, Sep 12, 2020 at 10:14 PM Bob kb8tq kb8tq@n1k.org wrote:
Hi
The unit your link points to is the “classic” TBolt that showed up in
volume a bit
over a decade ago. That one is packaged with a (noisy) switching power
supply.
There is no easy way to tell what era it is from. Indeed if it is from the
1990’s the
OCXO will not be quite as good as what’s in one from > 2002. Since the
unit normally
operates locked, that’s not a real big deal.
Since Trimble used “Thunderbolt” for a whole line of parts and was not
very good
at labeling devices, there is no easy way to refer to this or that
specific device. That
makes coming up with a list of units a bit hard. For example - what’s the
model number
of the part in that eBay listing ? ( No I don’t know either …. :) ).
Bob
On Sep 12, 2020, at 12:28 PM, Adrian Godwin artgodwin@gmail.com wrote:
I've been looking for a thunderbolt for a while, mostly because they're
such a benchmark within this group. I missed the period when they were
more
commonly available and have mostly found ones that were a lot more
expensive, or had a large delivery cost due to being in Australia.
Some cheaper examples turned up on UK ebay. These aren't the classic
time-nuts ones but presumably look fairly similar to Lady Heather.
https://www.ebay.co.uk/itm/154078750287
I haven't been able to find a comparison of the various thunderbolt
models.
Is there a description somewhere ? I've found a few hints about this one
I think it's a single 48V supply and possibly a lower-quality OCXO.
On Sat, Sep 12, 2020 at 5:14 PM Kevin Schuchmann kschuchm@gmail.com
wrote:
Matthias,
You are correct, having spent hours looking at all the stable areas
and cooling and heating the gpsdo I find that when it reports that it is
116.25 F then it is stable.
So now I guess I need to figure out why it is so picky. I will measure
the current draw from a cold start and see if I see the oven warming up
and then stabilizing and then heat and cool it and see how it reacts,
and also look at the electronics and see if an area is overly sensitive
to temperature.
Thanks
Kevin
On 9/11/2020 2:26 AM, Matthias Welwarsky wrote:
On Freitag, 11. September 2020 01:08:09 CEST Kevin Schuchmann wrote:
Guess my image didn't make it, I will add it as an attachment this
time.
The temperature curve seems to show some correlation to what is
happening with
the DAC. Seems that the DAC and OSC jumps are mostly during periods of
some
thermal perturbation.
Regards,
Matthias
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com