I have been off the list for a while, but noticed that there where
some posts on the LTZ1000 and LM399 about a month ago.
I just have a comment; the requirements on these two references
(standard datasheet implementations) are very much opposites:
LTZ1000 requires very stable resistors but is practically immune to
supply voltage variations.
LM399 does not require high stability resistors but requirements on
stable supply voltage are high.
On the $ of one LTZ1000 board, you can get a number of LM399
paralelled & averaged.
But one LTZ1000 runs on less power than one LM399
my 2c
Am 10.12.2014 um 16:20 schrieb Jan Fredriksson:
But one LTZ1000 runs on less power than one LM399
Hello,
I cannot agree to the last statement.
It mainly depends if you use LTZ1000 or LTZ1000A and temperature
setpoint and thermal isolation.
Just for comparison:
My two LTZ1000A with 50 deg C setpoint are using 20-21 mA each at 23 deg
C and 14V stabilized supply (from 12 NiMH cells = 15V battery)
My LM399#3 uses 30mA at room temp and 9.35V stabilized supply (from 8
NiMH cells = 10V battery)
so total power is around 300mW in both cases.
with best regards
Andreas