DJ
Didier Juges
Sun, Oct 29, 2006 5:31 AM
I moved the GPS antenna as high as I could in the shack (instead of on a
shelf at eye level) and it seems to have significantly improved the signal.
Here is the latest plot of the HP10811 against the GPS, unlocked.
http://www.ko4bb.com/Test_Equipment/HP10811-GPS-10-29.png
(sorry the screen shot is cropped at the bottom)
and the data file is at
http://www.ko4bb.com/Test_Equipment/HP10811-GPS-10-29.dat
Didier KO4BB
I moved the GPS antenna as high as I could in the shack (instead of on a
shelf at eye level) and it seems to have significantly improved the signal.
Here is the latest plot of the HP10811 against the GPS, unlocked.
http://www.ko4bb.com/Test_Equipment/HP10811-GPS-10-29.png
(sorry the screen shot is cropped at the bottom)
and the data file is at
http://www.ko4bb.com/Test_Equipment/HP10811-GPS-10-29.dat
Didier KO4BB
RK
Rob Kimberley
Sun, Oct 29, 2006 7:39 AM
Interesting to know what obstructions you cleared by moving it higher. May
have been getting some multipath from something on the antenna's apparent
horizon.
Rob K
-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Didier Juges
Sent: 29 October 2006 05:32
To: !Time Nuts
Subject: [time-nuts] More Allan deviation data
I moved the GPS antenna as high as I could in the shack (instead of on a
shelf at eye level) and it seems to have significantly improved the signal.
Here is the latest plot of the HP10811 against the GPS, unlocked.
http://www.ko4bb.com/Test_Equipment/HP10811-GPS-10-29.png
(sorry the screen shot is cropped at the bottom) and the data file is at
http://www.ko4bb.com/Test_Equipment/HP10811-GPS-10-29.dat
Didier KO4BB
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
Interesting to know what obstructions you cleared by moving it higher. May
have been getting some multipath from something on the antenna's apparent
horizon.
Rob K
-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Didier Juges
Sent: 29 October 2006 05:32
To: !Time Nuts
Subject: [time-nuts] More Allan deviation data
I moved the GPS antenna as high as I could in the shack (instead of on a
shelf at eye level) and it seems to have significantly improved the signal.
Here is the latest plot of the HP10811 against the GPS, unlocked.
http://www.ko4bb.com/Test_Equipment/HP10811-GPS-10-29.png
(sorry the screen shot is cropped at the bottom) and the data file is at
http://www.ko4bb.com/Test_Equipment/HP10811-GPS-10-29.dat
Didier KO4BB
_______________________________________________
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
UB
Ulrich Bangert
Sun, Oct 29, 2006 11:07 AM
Hi Didier,
you may use the DRIFT section of the graphic editor of PLOTTER to remove
a phase drift due to a constant frequency offset as seen in this data,
with the result that you get more "visual resolution" out of it, as in
the attached PDF.
73 de Ulrich, DF6JB
-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Didier Juges
Gesendet: Sonntag, 29. Oktober 2006 06:32
An: !Time Nuts
Betreff: [time-nuts] More Allan deviation data
I moved the GPS antenna as high as I could in the shack
(instead of on a
shelf at eye level) and it seems to have significantly
improved the signal.
Here is the latest plot of the HP10811 against the GPS, unlocked.
Hi Didier,
you may use the DRIFT section of the graphic editor of PLOTTER to remove
a phase drift due to a constant frequency offset as seen in this data,
with the result that you get more "visual resolution" out of it, as in
the attached PDF.
73 de Ulrich, DF6JB
> -----Ursprüngliche Nachricht-----
> Von: time-nuts-bounces@febo.com
> [mailto:time-nuts-bounces@febo.com] Im Auftrag von Didier Juges
> Gesendet: Sonntag, 29. Oktober 2006 06:32
> An: !Time Nuts
> Betreff: [time-nuts] More Allan deviation data
>
>
> I moved the GPS antenna as high as I could in the shack
> (instead of on a
> shelf at eye level) and it seems to have significantly
> improved the signal.
>
> Here is the latest plot of the HP10811 against the GPS, unlocked.
>
http://www.ko4bb.com/Test_Equipment/HP10811-GPS-10-29.png
(sorry the screen shot is cropped at the bottom)
and the data file is at
http://www.ko4bb.com/Test_Equipment/HP10811-GPS-10-29.dat
Didier KO4BB
_______________________________________________
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
DJ
Didier Juges
Sun, Oct 29, 2006 12:41 PM
I had to stack up frequency counters on top of frequency counters (to
measure time difference and each input frequency simultaneously -> 3
counters) and the last one on the top, a HP 5334B, was fairly high and
just about level with the antenna, about 2 feet away. I found out that
if I stood in line with the antenna and the counter, my shadow would
significantly affect the cleanliness of the 1 PPS, so I moved the
antenna just under the ceiling and the problem went away.
Amazing!
Didier
Rob Kimberley wrote:
Interesting to know what obstructions you cleared by moving it higher. May
have been getting some multipath from something on the antenna's apparent
horizon.
Rob K
I had to stack up frequency counters on top of frequency counters (to
measure time difference and each input frequency simultaneously -> 3
counters) and the last one on the top, a HP 5334B, was fairly high and
just about level with the antenna, about 2 feet away. I found out that
if I stood in line with the antenna and the counter, my shadow would
significantly affect the cleanliness of the 1 PPS, so I moved the
antenna just under the ceiling and the problem went away.
Amazing!
Didier
Rob Kimberley wrote:
> Interesting to know what obstructions you cleared by moving it higher. May
> have been getting some multipath from something on the antenna's apparent
> horizon.
>
> Rob K
DJ
Didier Juges
Sun, Oct 29, 2006 2:21 PM
Hi Didier,
you may use the DRIFT section of the graphic editor of PLOTTER to remove
a phase drift due to a constant frequency offset as seen in this data,
with the result that you get more "visual resolution" out of it, as in
the attached PDF.
73 de Ulrich, DF6JB
Ulrich,
I ran the OCXO with the new antenna location overnight, except that the
laptop went to sleep some time in the night, so I have lost several
hours of data, which may explain the sudden slope change in the plot
below (I need to save timestamps I guess). However, the remaining data
is good I think. There was no sudden noise, and about 5 phase jumps due
to the OCXO drifting during the observation period.
I downloaded the latest version of Plotter (10-27) and it's all there. I
was able to quickly remove the phase jumps when the OCXO drifted past
the divided down period, and then remove the remaining slow drift.
What a nice piece of software!
(even the menu now has a normal File->Open :-)
The latest plots are at the usual place:
http://www.ko4bb.com/Test_Equipment/HP10811-GPS_10-29-2.png
http://www.ko4bb.com/Test_Equipment/HP10811-GPS_10-29-2_Allan.png
Now that the data collection method seems to be getting close, I wonder
how I should use the data, and what it means.
The Allan deviation seems to taper around e-10 at 1000 seconds, then it
looks like it wants to keep getting lower.
As a reminder, this is the HP10811 against the GPS receiver. I am not
sure what I should expect?
Didier
PS: I am still planning to measure ambient temperature at the same time
as I collect TI, via the GPIB, using a HP 3478A voltmeter and a
thermistor. This way, I should be able to correlate frequency changes
and temperature changes. I assume I can tell Plotter to ignore the
temperature field or display it with its own scaling. Should I use a
specific character as field delimiter? (such as a comma, or just spaces?)
(In the short term, temperature data will be thermistor values, until I
write the code to convert in degrees. )
Also, if I use time stamps, should I use field delimiter between the
time stamp and the data?
What is the time stamps are discontinuous (as if the laptop falls asleep
again), will that mess up Plotter?
Ulrich Bangert wrote:
> Hi Didier,
>
> you may use the DRIFT section of the graphic editor of PLOTTER to remove
> a phase drift due to a constant frequency offset as seen in this data,
> with the result that you get more "visual resolution" out of it, as in
> the attached PDF.
>
> 73 de Ulrich, DF6JB
>
>
Ulrich,
I ran the OCXO with the new antenna location overnight, except that the
laptop went to sleep some time in the night, so I have lost several
hours of data, which may explain the sudden slope change in the plot
below (I need to save timestamps I guess). However, the remaining data
is good I think. There was no sudden noise, and about 5 phase jumps due
to the OCXO drifting during the observation period.
I downloaded the latest version of Plotter (10-27) and it's all there. I
was able to quickly remove the phase jumps when the OCXO drifted past
the divided down period, and then remove the remaining slow drift.
What a nice piece of software!
(even the menu now has a *normal* File->Open :-)
The latest plots are at the usual place:
http://www.ko4bb.com/Test_Equipment/HP10811-GPS_10-29-2.png
http://www.ko4bb.com/Test_Equipment/HP10811-GPS_10-29-2_Allan.png
Now that the data collection method seems to be getting close, I wonder
how I should use the data, and what it means.
The Allan deviation seems to taper around e-10 at 1000 seconds, then it
looks like it wants to keep getting lower.
As a reminder, this is the HP10811 against the GPS receiver. I am not
sure what I should expect?
Didier
PS: I am still planning to measure ambient temperature at the same time
as I collect TI, via the GPIB, using a HP 3478A voltmeter and a
thermistor. This way, I should be able to correlate frequency changes
and temperature changes. I assume I can tell Plotter to ignore the
temperature field or display it with its own scaling. Should I use a
specific character as field delimiter? (such as a comma, or just spaces?)
(In the short term, temperature data will be thermistor values, until I
write the code to convert in degrees. )
Also, if I use time stamps, should I use field delimiter between the
time stamp and the data?
What is the time stamps are discontinuous (as if the laptop falls asleep
again), will that mess up Plotter?
UB
Ulrich Bangert
Sun, Oct 29, 2006 7:46 PM
(even the menu now has a normal File->Open :-)
you take the credit for this!
73 de Ulrich, DF6JB
-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Didier Juges
Gesendet: Sonntag, 29. Oktober 2006 15:22
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] More Allan deviation data
Ulrich Bangert wrote:
Hi Didier,
you may use the DRIFT section of the graphic editor of PLOTTER to
remove a phase drift due to a constant frequency offset as seen in
this data, with the result that you get more "visual
of it, as in the attached PDF.
73 de Ulrich, DF6JB
Ulrich,
I ran the OCXO with the new antenna location overnight,
except that the
laptop went to sleep some time in the night, so I have lost several
hours of data, which may explain the sudden slope change in the plot
below (I need to save timestamps I guess). However, the
remaining data
is good I think. There was no sudden noise, and about 5 phase
jumps due
to the OCXO drifting during the observation period.
I downloaded the latest version of Plotter (10-27) and it's
all there. I
was able to quickly remove the phase jumps when the OCXO drifted past
the divided down period, and then remove the remaining slow drift.
What a nice piece of software!
(even the menu now has a normal File->Open :-)
The latest plots are at the usual place:
http://www.ko4bb.com/Test_Equipment/HP10811-> GPS_10-29-2.png
http://www.ko4bb.com/Test_Equipment/HP10811-GPS_10-29-2_Allan.png
Now that the data collection method seems to be getting close, I wonder
how I should use the data, and what it means.
The Allan deviation seems to taper around e-10 at 1000 seconds, then it
looks like it wants to keep getting lower.
As a reminder, this is the HP10811 against the GPS receiver. I am not
sure what I should expect?
Didier
PS: I am still planning to measure ambient temperature at the same time
as I collect TI, via the GPIB, using a HP 3478A voltmeter and a
thermistor. This way, I should be able to correlate frequency changes
and temperature changes. I assume I can tell Plotter to ignore the
temperature field or display it with its own scaling. Should I use a
specific character as field delimiter? (such as a comma, or just
spaces?)
(In the short term, temperature data will be thermistor values, until I
write the code to convert in degrees. )
Also, if I use time stamps, should I use field delimiter between the
time stamp and the data?
What is the time stamps are discontinuous (as if the laptop falls asleep
again), will that mess up Plotter?
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
Didier,
> (even the menu now has a *normal* File->Open :-)
you take the credit for this!
73 de Ulrich, DF6JB
> -----Ursprüngliche Nachricht-----
> Von: time-nuts-bounces@febo.com
> [mailto:time-nuts-bounces@febo.com] Im Auftrag von Didier Juges
> Gesendet: Sonntag, 29. Oktober 2006 15:22
> An: Discussion of precise time and frequency measurement
> Betreff: Re: [time-nuts] More Allan deviation data
>
>
> Ulrich Bangert wrote:
> > Hi Didier,
> >
> > you may use the DRIFT section of the graphic editor of PLOTTER to
> > remove a phase drift due to a constant frequency offset as seen in
> > this data, with the result that you get more "visual
> resolution" out
> > of it, as in the attached PDF.
> >
> > 73 de Ulrich, DF6JB
> >
> >
> Ulrich,
>
> I ran the OCXO with the new antenna location overnight,
> except that the
> laptop went to sleep some time in the night, so I have lost several
> hours of data, which may explain the sudden slope change in the plot
> below (I need to save timestamps I guess). However, the
> remaining data
> is good I think. There was no sudden noise, and about 5 phase
> jumps due
> to the OCXO drifting during the observation period.
>
> I downloaded the latest version of Plotter (10-27) and it's
> all there. I
> was able to quickly remove the phase jumps when the OCXO drifted past
> the divided down period, and then remove the remaining slow drift.
>
> What a nice piece of software!
> (even the menu now has a *normal* File->Open :-)
>
> The latest plots are at the usual place:
> http://www.ko4bb.com/Test_Equipment/HP10811-> GPS_10-29-2.png
>
>
http://www.ko4bb.com/Test_Equipment/HP10811-GPS_10-29-2_Allan.png
Now that the data collection method seems to be getting close, I wonder
how I should use the data, and what it means.
The Allan deviation seems to taper around e-10 at 1000 seconds, then it
looks like it wants to keep getting lower.
As a reminder, this is the HP10811 against the GPS receiver. I am not
sure what I should expect?
Didier
PS: I am still planning to measure ambient temperature at the same time
as I collect TI, via the GPIB, using a HP 3478A voltmeter and a
thermistor. This way, I should be able to correlate frequency changes
and temperature changes. I assume I can tell Plotter to ignore the
temperature field or display it with its own scaling. Should I use a
specific character as field delimiter? (such as a comma, or just
spaces?)
(In the short term, temperature data will be thermistor values, until I
write the code to convert in degrees. )
Also, if I use time stamps, should I use field delimiter between the
time stamp and the data?
What is the time stamps are discontinuous (as if the laptop falls asleep
again), will that mess up Plotter?
_______________________________________________
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
DB
Dr Bruce Griffiths
Sun, Oct 29, 2006 10:27 PM
Hi Didier,
you may use the DRIFT section of the graphic editor of PLOTTER to remove
a phase drift due to a constant frequency offset as seen in this data,
with the result that you get more "visual resolution" out of it, as in
the attached PDF.
73 de Ulrich, DF6JB
Ulrich,
I ran the OCXO with the new antenna location overnight, except that the
laptop went to sleep some time in the night, so I have lost several
hours of data, which may explain the sudden slope change in the plot
below (I need to save timestamps I guess). However, the remaining data
is good I think. There was no sudden noise, and about 5 phase jumps due
to the OCXO drifting during the observation period.
I downloaded the latest version of Plotter (10-27) and it's all there. I
was able to quickly remove the phase jumps when the OCXO drifted past
the divided down period, and then remove the remaining slow drift.
What a nice piece of software!
(even the menu now has a normal File->Open :-)
The latest plots are at the usual place:
http://www.ko4bb.com/Test_Equipment/HP10811-GPS_10-29-2.png
http://www.ko4bb.com/Test_Equipment/HP10811-GPS_10-29-2_Allan.png
Now that the data collection method seems to be getting close, I wonder
how I should use the data, and what it means.
The Allan deviation seems to taper around e-10 at 1000 seconds, then it
looks like it wants to keep getting lower.
As a reminder, this is the HP10811 against the GPS receiver. I am not
sure what I should expect?
Didier
PS: I am still planning to measure ambient temperature at the same time
as I collect TI, via the GPIB, using a HP 3478A voltmeter and a
thermistor. This way, I should be able to correlate frequency changes
and temperature changes. I assume I can tell Plotter to ignore the
temperature field or display it with its own scaling. Should I use a
specific character as field delimiter? (such as a comma, or just spaces?)
(In the short term, temperature data will be thermistor values, until I
write the code to convert in degrees. )
Also, if I use time stamps, should I use field delimiter between the
time stamp and the data?
What is the time stamps are discontinuous (as if the laptop falls asleep
again), will that mess up Plotter?
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
Didier
Unless you have a particularly poor 10811 it is likely that the measured
Allan deviation is limited by your the instability of your GPS receiver
PPS output.
With a good GPS timing receiver setup you should expect an Allan
variance perhaps 5x lower than this as illustrated in Ulrich's article
for the AMSAT journal.
Bruce
Didier Juges wrote:
> Ulrich Bangert wrote:
>
>> Hi Didier,
>>
>> you may use the DRIFT section of the graphic editor of PLOTTER to remove
>> a phase drift due to a constant frequency offset as seen in this data,
>> with the result that you get more "visual resolution" out of it, as in
>> the attached PDF.
>>
>> 73 de Ulrich, DF6JB
>>
>>
>>
> Ulrich,
>
> I ran the OCXO with the new antenna location overnight, except that the
> laptop went to sleep some time in the night, so I have lost several
> hours of data, which may explain the sudden slope change in the plot
> below (I need to save timestamps I guess). However, the remaining data
> is good I think. There was no sudden noise, and about 5 phase jumps due
> to the OCXO drifting during the observation period.
>
> I downloaded the latest version of Plotter (10-27) and it's all there. I
> was able to quickly remove the phase jumps when the OCXO drifted past
> the divided down period, and then remove the remaining slow drift.
>
> What a nice piece of software!
> (even the menu now has a *normal* File->Open :-)
>
> The latest plots are at the usual place:
> http://www.ko4bb.com/Test_Equipment/HP10811-GPS_10-29-2.png
> http://www.ko4bb.com/Test_Equipment/HP10811-GPS_10-29-2_Allan.png
>
> Now that the data collection method seems to be getting close, I wonder
> how I should use the data, and what it means.
>
> The Allan deviation seems to taper around e-10 at 1000 seconds, then it
> looks like it wants to keep getting lower.
>
> As a reminder, this is the HP10811 against the GPS receiver. I am not
> sure what I should expect?
>
> Didier
>
> PS: I am still planning to measure ambient temperature at the same time
> as I collect TI, via the GPIB, using a HP 3478A voltmeter and a
> thermistor. This way, I should be able to correlate frequency changes
> and temperature changes. I assume I can tell Plotter to ignore the
> temperature field or display it with its own scaling. Should I use a
> specific character as field delimiter? (such as a comma, or just spaces?)
>
> (In the short term, temperature data will be thermistor values, until I
> write the code to convert in degrees. )
>
> Also, if I use time stamps, should I use field delimiter between the
> time stamp and the data?
>
> What is the time stamps are discontinuous (as if the laptop falls asleep
> again), will that mess up Plotter?
>
>
>
> _______________________________________________
> time-nuts mailing list
> time-nuts@febo.com
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>
>
Didier
Unless you have a particularly poor 10811 it is likely that the measured
Allan deviation is limited by your the instability of your GPS receiver
PPS output.
With a good GPS timing receiver setup you should expect an Allan
variance perhaps 5x lower than this as illustrated in Ulrich's article
for the AMSAT journal.
Bruce
DJ
Didier Juges
Sun, Oct 29, 2006 11:32 PM
Dr Bruce Griffiths wrote:
Didier
Unless you have a particularly poor 10811 it is likely that the measured
Allan deviation is limited by the instability of your GPS receiver
PPS output.
With a good GPS timing receiver setup you should expect an Allan
variance perhaps 5x lower than this as illustrated in Ulrich's article
for the AMSAT journal.
Bruce
Thank you Bruce, that is the kind of information I was looking for. The
Jupiter will have to do for now, I have two of the same model. I may try
a better antenna though.
I got a link to the gpskit.nl web page, with a lot of information on the
Jupiter, through an off-list message, and I will experiment with the
settings in the receiver (mask angle primarily)
It seems all I have left to do is find the best way to close the loop
(!) and try again against another 10811. I also still have a lot of
experimentation to do over temperature, and check the stability of my
distribution amplifier.
I have learned a lot in the last few days, thanks to everyone on
time-nuts who took the time to explain the details to me.
I have a lot left to learn and experiment with, but at least I am aiming
in the right direction :-)
Didier KO4BB
Dr Bruce Griffiths wrote:
> Didier
>
> Unless you have a particularly poor 10811 it is likely that the measured
> Allan deviation is limited by the instability of your GPS receiver
> PPS output.
> With a good GPS timing receiver setup you should expect an Allan
> variance perhaps 5x lower than this as illustrated in Ulrich's article
> for the AMSAT journal.
>
> Bruce
>
>
Thank you Bruce, that is the kind of information I was looking for. The
Jupiter will have to do for now, I have two of the same model. I may try
a better antenna though.
I got a link to the gpskit.nl web page, with a lot of information on the
Jupiter, through an off-list message, and I will experiment with the
settings in the receiver (mask angle primarily)
It seems all I have left to do is find the best way to close the loop
(!) and try again against another 10811. I also still have a lot of
experimentation to do over temperature, and check the stability of my
distribution amplifier.
I have learned a lot in the last few days, thanks to everyone on
time-nuts who took the time to explain the details to me.
I have a lot left to learn and experiment with, but at least I am aiming
in the right direction :-)
Didier KO4BB