My rubidium frequency standard (Accubeat AR60A) finally got its nice box
with good cooling and some LED's showing the status.
While watching the unlock LED I noticed the AR60 say's it is locked,
about 4 or 5 minutes after cold start, when the frequency error jumps
below about 1e-8
It takes 2 or 3 more minutes of slow drift to become stable within
1e-10. Is this to be expected?
With this locking speed it is tempting to switch off the AR60A unless
needed for some measurement. Would daily switching on/off have any
impact on lifetime?
Hi
Yes, that’s pretty normal operation for any of the telecom Rb’s. They
also typically drift about 1x10^-9 over the first day of operation. It’s
best to leave them running all the time. Second best is to turn them
on a couple days before you need to use them.
Bob
On Dec 11, 2021, at 2:23 AM, Erik Kaashoek erik@kaashoek.com wrote:
My rubidium frequency standard (Accubeat AR60A) finally got its nice box with good cooling and some LED's showing the status.
While watching the unlock LED I noticed the AR60 say's it is locked, about 4 or 5 minutes after cold start, when the frequency error jumps below about 1e-8
It takes 2 or 3 more minutes of slow drift to become stable within 1e-10. Is this to be expected?
With this locking speed it is tempting to switch off the AR60A unless needed for some measurement. Would daily switching on/off have any impact on lifetime?
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
Totally agree with Bob. Same experience.
On Sat, Dec 11, 2021 at 8:12 AM Bob kb8tq kb8tq@n1k.org wrote:
Hi
Yes, that’s pretty normal operation for any of the telecom Rb’s. They
also typically drift about 1x10^-9 over the first day of operation. It’s
best to leave them running all the time. Second best is to turn them
on a couple days before you need to use them.
Bob
On Dec 11, 2021, at 2:23 AM, Erik Kaashoek erik@kaashoek.com wrote:
My rubidium frequency standard (Accubeat AR60A) finally got its nice box
with good cooling and some LED's showing the status.
While watching the unlock LED I noticed the AR60 say's it is locked,
about 4 or 5 minutes after cold start, when the frequency error jumps below
about 1e-8
It takes 2 or 3 more minutes of slow drift to become stable within
1e-10. Is this to be expected?
With this locking speed it is tempting to switch off the AR60A unless
needed for some measurement. Would daily switching on/off have any impact
on lifetime?
send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
hi Paul, Bob,
Is there a significant difference between these telco-surplus Rb and the more 'fancy' ones in this regard?
I happen to own an Efratom MRK-HLN, which, I presume, is also more geared towards "always on". So, I wonder would the MRK be better or worse compared to a telco-surplus unit? Arguably a telco-surplus unit is also of the always-on variety or?
best,
Wilko
(newbie trying to get a feel for this)
On 11 Dec 2021, at 16:27, paul swed paulswedb@gmail.com wrote:
Totally agree with Bob. Same experience.
On Sat, Dec 11, 2021 at 8:12 AM Bob kb8tq kb8tq@n1k.org wrote:
Hi
Yes, that’s pretty normal operation for any of the telecom Rb’s. They
also typically drift about 1x10^-9 over the first day of operation. It’s
best to leave them running all the time. Second best is to turn them
on a couple days before you need to use them.
Bob
On Dec 11, 2021, at 2:23 AM, Erik Kaashoek erik@kaashoek.com wrote:
My rubidium frequency standard (Accubeat AR60A) finally got its nice box
with good cooling and some LED's showing the status.
While watching the unlock LED I noticed the AR60 say's it is locked,
about 4 or 5 minutes after cold start, when the frequency error jumps below
about 1e-8
It takes 2 or 3 more minutes of slow drift to become stable within
1e-10. Is this to be expected?
With this locking speed it is tempting to switch off the AR60A unless
needed for some measurement. Would daily switching on/off have any impact
on lifetime?
send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
Wilco
Great to hear from you. Here is my 10 cents. Most of the things we happen
to find are generally of the telco grade. The FRS as an example is a nicely
built unit that used to be had for as little as $25. I picked a few up.
Well maybe 8. These had all been in service (approx 1994) and so after a
while they tended to suffer the plating out of the RB on the lamp. Ending
the useful life. Thanks to time-nuts its been shown you can melt the RB off
the lamp and you are back in business. For me at least another 5 years. The
better units most likely have better reserve in the lamp. Maybe easier to
repair and be stabilized perhaps a bit better.
I would trust Bob's thoughts, thats why mine is worth 10 cents.
Regards
Paul
WB8TSL
On Sat, Dec 11, 2021 at 1:37 PM Wilko Bulte wkb@xs4all.nl wrote:
hi Paul, Bob,
Is there a significant difference between these telco-surplus Rb and the
more 'fancy' ones in this regard?
I happen to own an Efratom MRK-HLN, which, I presume, is also more geared
towards "always on". So, I wonder would the MRK be better or worse compared
to a telco-surplus unit? Arguably a telco-surplus unit is also of the
always-on variety or?
best,
Wilko
(newbie trying to get a feel for this)
On 11 Dec 2021, at 16:27, paul swed paulswedb@gmail.com wrote:
Totally agree with Bob. Same experience.
On Sat, Dec 11, 2021 at 8:12 AM Bob kb8tq kb8tq@n1k.org wrote:
Hi
Yes, that’s pretty normal operation for any of the telecom Rb’s. They
also typically drift about 1x10^-9 over the first day of operation. It’s
best to leave them running all the time. Second best is to turn them
on a couple days before you need to use them.
Bob
On Dec 11, 2021, at 2:23 AM, Erik Kaashoek erik@kaashoek.com wrote:
My rubidium frequency standard (Accubeat AR60A) finally got its nice
box
with good cooling and some LED's showing the status.
While watching the unlock LED I noticed the AR60 say's it is locked,
about 4 or 5 minutes after cold start, when the frequency error jumps
below
about 1e-8
It takes 2 or 3 more minutes of slow drift to become stable within
1e-10. Is this to be expected?
With this locking speed it is tempting to switch off the AR60A unless
needed for some measurement. Would daily switching on/off have any
impact
on lifetime?
send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe
send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe
send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
Hi
The Efratom “big” Rb’s are sort of an early version of the telco devices.
Their design dates back well before the smaller units came out. There
also are very similar devices made by EG&G ( we wanted a part of that
market, back in the day ….).
Bob
On Dec 11, 2021, at 10:59 AM, Wilko Bulte wkb@xs4all.nl wrote:
hi Paul, Bob,
Is there a significant difference between these telco-surplus Rb and the more 'fancy' ones in this regard?
I happen to own an Efratom MRK-HLN, which, I presume, is also more geared towards "always on". So, I wonder would the MRK be better or worse compared to a telco-surplus unit? Arguably a telco-surplus unit is also of the always-on variety or?
best,
Wilko
(newbie trying to get a feel for this)
On 11 Dec 2021, at 16:27, paul swed paulswedb@gmail.com wrote:
Totally agree with Bob. Same experience.
On Sat, Dec 11, 2021 at 8:12 AM Bob kb8tq kb8tq@n1k.org wrote:
Hi
Yes, that’s pretty normal operation for any of the telecom Rb’s. They
also typically drift about 1x10^-9 over the first day of operation. It’s
best to leave them running all the time. Second best is to turn them
on a couple days before you need to use them.
Bob
On Dec 11, 2021, at 2:23 AM, Erik Kaashoek erik@kaashoek.com wrote:
My rubidium frequency standard (Accubeat AR60A) finally got its nice box
with good cooling and some LED's showing the status.
While watching the unlock LED I noticed the AR60 say's it is locked,
about 4 or 5 minutes after cold start, when the frequency error jumps below
about 1e-8
It takes 2 or 3 more minutes of slow drift to become stable within
1e-10. Is this to be expected?
With this locking speed it is tempting to switch off the AR60A unless
needed for some measurement. Would daily switching on/off have any impact
on lifetime?
send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.