Hi Bob

I view the master/slave situation the same way you do, but only  commented
because Lucent effectively sets things up the other way round so just
wanted to be sure which way Paul was considering it.

As regards, the "faking", yes, a single 15 way plug containing just  a link
and a resistor is exactly what I use now.

I realise the RS422 fudge isn't ideal but certainly very handy for some
quick tests.

I've got two or three RS422 PCI interfaces but hadn't previously had any
spare slots, now I've got a couple of lovely Magma 13 way PCI expansion units
but need to hack them about so the ports are at the front.
Whoever decided rack mount PC kit should be built with all the  connections
at the rear must have been really nuts, and it was too long ago  to use
"corridors" of racks as a viable excuse, someone somewhere just fitted  ears to
desk top style cases and left the rest of us to get on with  it..

One of the best things I ever did was to buy a pair of "back to front"
rackmount PC cases with all ports and cards available at the front, trouble is
now I want everything else to match, and most test gear has just the same
issues!

All good fun, and certainly never a dull moment:-)

Regards

Nigel
GM8PZR

In a message dated 06/11/2014 12:04:47 GMT Standard Time, kb8tq@n1k.org
writes:

Hi

On Nov 6, 2014, at 3:44 AM, GandalfG8--- via  time-nuts

Hi Paul

Prime question is what are you calling the "slave" unit?

  1. Slave in the sense that it’s slaved to the GPS in the other box. Unless
    we  find a serial link from the Ref-0 back into the Ref-1, that’s all it’s
    doing.
  2. It fires up and the Ref-1 box gets all organized and locked up.  It goes
    “ok” first, then the Ref-0 goes “ok” (locks to GPS). In that sense,  the
    Ref-1 is the master.

My guess is that they default to output from  the Ref-0 rather than Ref-1
because it’s the one more likely to fail (more  single points to go wrong).

In normal use Lucent treats the Ref-0  unit as the master and Ref-1 as

the

slave, but it's Ref-1 that  contains the GPS module. Bit confusing

Yes it is confusing - sorry  about that

but if
you've  got Ref-1 with the GPS  module then you're in business:-)

It's turned out the  transistor delay circuit isn't needed after all and

the

wiring can be  simplified too.

All that's necessary to run the Ref-1 stand  alone is to ground pins 2

and 3

on the J5 interface connector and  there's a couple of convenient ground
pins  already on that  connector, 8 and 13.
In normal use pin3 is tied hard to ground,  presumably via direct link

from

the other module, so that can  just be a wire link to either pin 8 or 13.
Pin 2 is normally held at a  logic low level so although a direct  ground
seems to work ok  for that too, and no damage noted so far, I prefer to

play

it  safe and ground that one to the other ground pin via a 470ohm

resistor,

which also works fine.
That's it.

So a single 15 pin  connector with one jumper and one resistor (or jumper)
plugged into the  interface socket.

The LEDS should flash at start up,  followed by "No GPS" and "Fault" on
solid whilst the unit goes  through acquisition and lock, this can take

an hour

or so, and  then those LEDs should go off and the "On" light should  be
solid.
At this point the 15MHz and 1PPS outputs are enabled.

The J8 diagnostics connector is enabled from start up and SatStat  can

be

used to monitor progress.
The RS422 3 wire fudge  suggested by Stu Cobb works fine for me, J8 pin 7

to

PC 9 way comm  port pin 5, J8 pin 8 to comm port pin 3, J8 pin 9 to comm
port  pin  2.

Works, but if you have an old style +/- 12 V RS-232  signal, things might
or might not work forever. A lot depends on how they have  protected the
chips. They expect a 1.5 to 3.5V signal on the input, so even  +/-5 is a bit
exciting. Full blown RS-422 interfaces are sub $15  gizmos.

Bob

Regards

Nigel
GM8PZR

In a message dated 06/11/2014  01:24:33 GMT Standard Time,
paulswedb@gmail.com writes:

I have  been following the threads on the KS-24361 some of the  longest I
have ever  seen on time nuts.
So I have a very  basic question.
There was a 2  transistor delay circuit by Arthur  way back. I believe

this

is  needed.
But then the pin  numbers seem to get a bit mixed up and there are  some
jumpers  with several comments.
Is there a clear guidance on what to  do?  I do not have the slave unit.
Power is not an issue.
Thanks  in  advance for your help.
Must say the units very  nice.

Regards
Paul
WB8TSL


time-nuts  mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and  follow the  instructions there.


time-nuts mailing list  -- time-nuts@febo.com
To unsubscribe, go to

and follow the  instructions there.

Hi Bob I view the master/slave situation the same way you do, but only commented because Lucent effectively sets things up the other way round so just wanted to be sure which way Paul was considering it. As regards, the "faking", yes, a single 15 way plug containing just a link and a resistor is exactly what I use now. I realise the RS422 fudge isn't ideal but certainly very handy for some quick tests. I've got two or three RS422 PCI interfaces but hadn't previously had any spare slots, now I've got a couple of lovely Magma 13 way PCI expansion units but need to hack them about so the ports are at the front. Whoever decided rack mount PC kit should be built with all the connections at the rear must have been really nuts, and it was too long ago to use "corridors" of racks as a viable excuse, someone somewhere just fitted ears to desk top style cases and left the rest of us to get on with it.. One of the best things I ever did was to buy a pair of "back to front" rackmount PC cases with all ports and cards available at the front, trouble is now I want everything else to match, and most test gear has just the same issues! All good fun, and certainly never a dull moment:-) Regards Nigel GM8PZR In a message dated 06/11/2014 12:04:47 GMT Standard Time, kb8tq@n1k.org writes: Hi > On Nov 6, 2014, at 3:44 AM, GandalfG8--- via time-nuts <time-nuts@febo.com> wrote: > > Hi Paul > > Prime question is what are you calling the "slave" unit? 1) Slave in the sense that it’s slaved to the GPS in the other box. Unless we find a serial link from the Ref-0 back into the Ref-1, that’s all it’s doing. 2) It fires up and the Ref-1 box gets all organized and locked up. It goes “ok” first, then the Ref-0 goes “ok” (locks to GPS). In that sense, the Ref-1 is the master. My guess is that they default to output from the Ref-0 rather than Ref-1 because it’s the one more likely to fail (more single points to go wrong). > In normal use Lucent treats the Ref-0 unit as the master and Ref-1 as the > slave, but it's Ref-1 that contains the GPS module. Bit confusing Yes it is confusing - sorry about that > but if > you've got Ref-1 with the GPS module then you're in business:-) > > It's turned out the transistor delay circuit isn't needed after all and the > wiring can be simplified too. > > All that's necessary to run the Ref-1 stand alone is to ground pins 2 and 3 > on the J5 interface connector and there's a couple of convenient ground > pins already on that connector, 8 and 13. > In normal use pin3 is tied hard to ground, presumably via direct link from > the other module, so that can just be a wire link to either pin 8 or 13. > Pin 2 is normally held at a logic low level so although a direct ground > seems to work ok for that too, and no damage noted so far, I prefer to play > it safe and ground that one to the other ground pin via a 470ohm resistor, > which also works fine. > That's it. So a single 15 pin connector with one jumper and one resistor (or jumper) plugged into the interface socket. > > The LEDS should flash at start up, followed by "No GPS" and "Fault" on > solid whilst the unit goes through acquisition and lock, this can take an hour > or so, and then those LEDs should go off and the "On" light should be > solid. > At this point the 15MHz and 1PPS outputs are enabled. > > The J8 diagnostics connector is enabled from start up and SatStat can be > used to monitor progress. > The RS422 3 wire fudge suggested by Stu Cobb works fine for me, J8 pin 7 to > PC 9 way comm port pin 5, J8 pin 8 to comm port pin 3, J8 pin 9 to comm > port pin 2. Works, but if you have an old style +/- 12 V RS-232 signal, things might or might not work forever. A lot depends on how they have protected the chips. They expect a 1.5 to 3.5V signal on the input, so even +/-5 is a bit exciting. Full blown RS-422 interfaces are sub $15 gizmos. Bob > > Regards > > Nigel > GM8PZR > > > > In a message dated 06/11/2014 01:24:33 GMT Standard Time, > paulswedb@gmail.com writes: > > I have been following the threads on the KS-24361 some of the longest I > have ever seen on time nuts. > So I have a very basic question. > There was a 2 transistor delay circuit by Arthur way back. I believe this > is needed. > But then the pin numbers seem to get a bit mixed up and there are some > jumpers with several comments. > Is there a clear guidance on what to do? I do not have the slave unit. > Power is not an issue. > Thanks in advance for your help. > Must say the units very nice. > > Regards > Paul > WB8TSL > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BC
Bob Camp
Fri, Nov 7, 2014 12:07 AM

Hi

Ok, I got to spend a little quality time with my pair today. I mostly poked at the Diag port with a terminal program.

Both of the units seem talk over the Diag port regardless of the mode they are in (master / slave / fault ). Both respond to a sub-set of the Z3801 SCPI commands (= not all commands work). There are a few enhancements where a query to a “node” gives a reply (which it should not if it’s “node only”). All of it works ok at 9600 baud 8N1. All of it comes back with an error message (E-xxx) after a valid command response. It may not like the terminal program sending cr/lf or something like that.

By far the most useful thing to type is :SYST:STAT?
That executes the system:status request and gives you a nice full screen of information. It shows you about 95% of the useful information all on one page. About the only major thing missing is the DAC value. That’s available with another command out of the Z3801 list. It does take a while to compose and transmit the screen, so it’s not the best way to to this forever and ever.

The data from the request is the same on both boxes. Same everything except one shows as standby etc. The interesting details are things like elevation mask and cable delay. Why interesting? You can set them with discrete commands. You can also query the data with the same commands to make sure the box “took” what you sent.

This allows experimentation with changing the elevation mask on each box (or just simply doing a discrete query for the mask on each box). The Ref-0 / slave will happily change things around with discrete commands. None of them are reflected in the :SYST:STAT? listing on the Ref-0 or Ref-1. When you look at the Ref-1 / gps box, it’s discrete query information matches the :SYST:STAT? screen on both boxes. When you change things on the Ref-1 box both screens follow that data.

—————

When you pull the antenna on the gps box, it immediately (almost) goes into gps fault led flashing mode. The slave box happily sits there for a while (say a minute) with it’s gps led saying all is well. When you plug the antenna back in, the process is reversed. The gps box goes out of flash mode right away. The led goes out in a bit. The led on the slave box does not follow for a minute or so.

—————

All of this suggests to me that the Ref-0 / slave does not talk back to the other box via serial. It also suggests to me that all it “sees” from the Ref-1/gps box are the duplicates of the serial string out of the Motorola gps module. The Ref-0 just figures out what’s happening by watching those strings as they roll by.

Hope that makes sense….

——————

If that’s all true, it should be “easy” to put dummy gps sentences into the Ref-0 interface port and a pps from something else. Once it sees stuff that it likes, it will start locking up to that pps. Cheap micro and just about any modern gps = fancy new GPSDO. Quick / easy / not much work. A little more complex if you decide to translate sawtooth data. That of course assumes these boxes do anything with sawtooth ...

Bob

On Nov 6, 2014, at 7:52 AM, GandalfG8--- via time-nuts time-nuts@febo.com wrote:

Hi Bob

I view the master/slave situation the same way you do, but only  commented
because Lucent effectively sets things up the other way round so just
wanted to be sure which way Paul was considering it.

As regards, the "faking", yes, a single 15 way plug containing just  a link
and a resistor is exactly what I use now.

I realise the RS422 fudge isn't ideal but certainly very handy for some
quick tests.

I've got two or three RS422 PCI interfaces but hadn't previously had any
spare slots, now I've got a couple of lovely Magma 13 way PCI expansion units
but need to hack them about so the ports are at the front.
Whoever decided rack mount PC kit should be built with all the  connections
at the rear must have been really nuts, and it was too long ago  to use
"corridors" of racks as a viable excuse, someone somewhere just fitted  ears to
desk top style cases and left the rest of us to get on with  it..

One of the best things I ever did was to buy a pair of "back to front"
rackmount PC cases with all ports and cards available at the front, trouble is
now I want everything else to match, and most test gear has just the same
issues!

All good fun, and certainly never a dull moment:-)

Regards

Nigel
GM8PZR

Hi Ok, I got to spend a little quality time with my pair today. I mostly poked at the Diag port with a terminal program. Both of the units seem talk over the Diag port regardless of the mode they are in (master / slave / fault ). Both respond to a sub-set of the Z3801 SCPI commands (= not all commands work). There are a few enhancements where a query to a “node” gives a reply (which it should not if it’s “node only”). All of it works ok at 9600 baud 8N1. All of it comes back with an error message (E-xxx) after a valid command response. It may not like the terminal program sending cr/lf or something like that. By far the most useful thing to type is :SYST:STAT? That executes the system:status request and gives you a nice full screen of information. It shows you about 95% of the useful information all on one page. About the only major thing missing is the DAC value. That’s available with another command out of the Z3801 list. It does take a while to compose and transmit the screen, so it’s not the best way to to this forever and ever. The data from the request is the same on both boxes. Same everything except one shows as standby etc. The interesting details are things like elevation mask and cable delay. Why interesting? You can set them with discrete commands. You can also query the data with the same commands to make sure the box “took” what you sent. This allows experimentation with changing the elevation mask on each box (or just simply doing a discrete query for the mask on each box). The Ref-0 / slave will happily change things around with discrete commands. None of them are reflected in the :SYST:STAT? listing on the Ref-0 or Ref-1. When you look at the Ref-1 / gps box, it’s discrete query information matches the :SYST:STAT? screen on both boxes. When you change things on the Ref-1 box both screens follow that data. ————— When you pull the antenna on the gps box, it immediately (almost) goes into gps fault led flashing mode. The slave box happily sits there for a while (say a minute) with it’s gps led saying all is well. When you plug the antenna back in, the process is reversed. The gps box goes out of flash mode right away. The led goes out in a bit. The led on the slave box does not follow for a minute or so. ————— All of this suggests to me that the Ref-0 / slave does not talk back to the other box via serial. It also suggests to me that all it “sees” from the Ref-1/gps box are the duplicates of the serial string out of the Motorola gps module. The Ref-0 just figures out what’s happening by watching those strings as they roll by. Hope that makes sense…. —————— If that’s all true, it should be “easy” to put dummy gps sentences into the Ref-0 interface port and a pps from something else. Once it sees stuff that it likes, it will start locking up to that pps. Cheap micro and just about any modern gps = fancy new GPSDO. Quick / easy / not much work. A little more complex if you decide to translate sawtooth data. That of course assumes these boxes do anything with sawtooth ... Bob > On Nov 6, 2014, at 7:52 AM, GandalfG8--- via time-nuts <time-nuts@febo.com> wrote: > > Hi Bob > > I view the master/slave situation the same way you do, but only commented > because Lucent effectively sets things up the other way round so just > wanted to be sure which way Paul was considering it. > > As regards, the "faking", yes, a single 15 way plug containing just a link > and a resistor is exactly what I use now. > > I realise the RS422 fudge isn't ideal but certainly very handy for some > quick tests. > > I've got two or three RS422 PCI interfaces but hadn't previously had any > spare slots, now I've got a couple of lovely Magma 13 way PCI expansion units > but need to hack them about so the ports are at the front. > Whoever decided rack mount PC kit should be built with all the connections > at the rear must have been really nuts, and it was too long ago to use > "corridors" of racks as a viable excuse, someone somewhere just fitted ears to > desk top style cases and left the rest of us to get on with it.. > > One of the best things I ever did was to buy a pair of "back to front" > rackmount PC cases with all ports and cards available at the front, trouble is > now I want everything else to match, and most test gear has just the same > issues! > > All good fun, and certainly never a dull moment:-) > > Regards > > Nigel > GM8PZR > > >
BS
Bob Stewart
Fri, Nov 7, 2014 5:02 AM

Hi Bob,
I've been using the Satstat program to send commands.  What I don't understand is why the slave will accept commands such as sat elevation angle and not give an error response.  It gives a response of "command complete", but it doesn't change it.  Or is that an indication that both units use the same firmware, and success is actually indicated by the resulting change of the angle?

Bob    From: Bob Camp kb8tq@n1k.org
To: GandalfG8@aol.com; Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Thursday, November 6, 2014 6:07 PM
Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions

Hi

Ok, I got to spend a little quality time with my pair today. I mostly poked at the Diag port with a terminal program.

Both of the units seem talk over the Diag port regardless of the mode they are in (master / slave / fault ). Both respond to a sub-set of the Z3801 SCPI commands (= not all commands work). There are a few enhancements where a query to a “node” gives a reply (which it should not if it’s “node only”). All of it works ok at 9600 baud 8N1. All of it comes back with an error message (E-xxx) after a valid command response. It may not like the terminal program sending cr/lf or something like that.

By far the most useful thing to type is :SYST:STAT?
That executes the system:status request and gives you a nice full screen of information. It shows you about 95% of the useful information all on one page. About the only major thing missing is the DAC value. That’s available with another command out of the Z3801 list. It does take a while to compose and transmit the screen, so it’s not the best way to to this forever and ever.

The data from the request is the same on both boxes. Same everything except one shows as standby etc. The interesting details are things like elevation mask and cable delay. Why interesting? You can set them with discrete commands. You can also query the data with the same commands to make sure the box “took” what you sent.

This allows experimentation with changing the elevation mask on each box (or just simply doing a discrete query for the mask on each box). The Ref-0 / slave will happily change things around with discrete commands. None of them are reflected in the :SYST:STAT? listing on the Ref-0 or Ref-1. When you look at the Ref-1 / gps box, it’s discrete query information matches the :SYST:STAT? screen on both boxes. When you change things on the Ref-1 box both screens follow that data.

—————

When you pull the antenna on the gps box, it immediately (almost) goes into gps fault led flashing mode. The slave box happily sits there for a while (say a minute) with it’s gps led saying all is well. When you plug the antenna back in, the process is reversed. The gps box goes out of flash mode right away. The led goes out in a bit. The led on the slave box does not follow for a minute or so.

—————

All of this suggests to me that the Ref-0 / slave does not talk back to the other box via serial. It also suggests to me that all it “sees” from the Ref-1/gps box are the duplicates of the serial string out of the Motorola gps module. The Ref-0 just figures out what’s happening by watching those strings as they roll by.

Hope that makes sense….

——————

If that’s all true, it should be “easy” to put dummy gps sentences into the Ref-0 interface port and a pps from something else. Once it sees stuff that it likes, it will start locking up to that pps. Cheap micro and just about any modern gps = fancy new GPSDO. Quick / easy / not much work. A little more complex if you decide to translate sawtooth data. That of course assumes these boxes do anything with sawtooth ...

Bob

On Nov 6, 2014, at 7:52 AM, GandalfG8--- via time-nuts time-nuts@febo.com wrote:

Hi Bob

I view the master/slave situation the same way you do, but only  commented
because Lucent effectively sets things up the other way round so just 
wanted to be sure which way Paul was considering it.

As regards, the "faking", yes, a single 15 way plug containing just  a link
and a resistor is exactly what I use now.

I realise the RS422 fudge isn't ideal but certainly very handy for some 
quick tests.

I've got two or three RS422 PCI interfaces but hadn't previously had any 
spare slots, now I've got a couple of lovely Magma 13 way PCI expansion units
but need to hack them about so the ports are at the front.
Whoever decided rack mount PC kit should be built with all the  connections
at the rear must have been really nuts, and it was too long ago  to use
"corridors" of racks as a viable excuse, someone somewhere just fitted  ears to
desk top style cases and left the rest of us to get on with  it..

One of the best things I ever did was to buy a pair of "back to front" 
rackmount PC cases with all ports and cards available at the front, trouble is 
now I want everything else to match, and most test gear has just the same 
issues!

All good fun, and certainly never a dull moment:-)

Regards

Nigel
GM8PZR


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Bob, I've been using the Satstat program to send commands.  What I don't understand is why the slave will accept commands such as sat elevation angle and not give an error response.  It gives a response of "command complete", but it doesn't change it.  Or is that an indication that both units use the same firmware, and success is actually indicated by the resulting change of the angle? Bob From: Bob Camp <kb8tq@n1k.org> To: GandalfG8@aol.com; Discussion of precise time and frequency measurement <time-nuts@febo.com> Sent: Thursday, November 6, 2014 6:07 PM Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions Hi Ok, I got to spend a little quality time with my pair today. I mostly poked at the Diag port with a terminal program. Both of the units seem talk over the Diag port regardless of the mode they are in (master / slave / fault ). Both respond to a sub-set of the Z3801 SCPI commands (= not all commands work). There are a few enhancements where a query to a “node” gives a reply (which it should not if it’s “node only”). All of it works ok at 9600 baud 8N1. All of it comes back with an error message (E-xxx) after a valid command response. It may not like the terminal program sending cr/lf or something like that. By far the most useful thing to type is :SYST:STAT? That executes the system:status request and gives you a nice full screen of information. It shows you about 95% of the useful information all on one page. About the only major thing missing is the DAC value. That’s available with another command out of the Z3801 list. It does take a while to compose and transmit the screen, so it’s not the best way to to this forever and ever. The data from the request is the same on both boxes. Same everything except one shows as standby etc. The interesting details are things like elevation mask and cable delay. Why interesting? You can set them with discrete commands. You can also query the data with the same commands to make sure the box “took” what you sent. This allows experimentation with changing the elevation mask on each box (or just simply doing a discrete query for the mask on each box). The Ref-0 / slave will happily change things around with discrete commands. None of them are reflected in the :SYST:STAT? listing on the Ref-0 or Ref-1. When you look at the Ref-1 / gps box, it’s discrete query information matches the :SYST:STAT? screen on both boxes. When you change things on the Ref-1 box both screens follow that data. ————— When you pull the antenna on the gps box, it immediately (almost) goes into gps fault led flashing mode. The slave box happily sits there for a while (say a minute) with it’s gps led saying all is well. When you plug the antenna back in, the process is reversed. The gps box goes out of flash mode right away. The led goes out in a bit. The led on the slave box does not follow for a minute or so. ————— All of this suggests to me that the Ref-0 / slave does not talk back to the other box via serial. It also suggests to me that all it “sees” from the Ref-1/gps box are the duplicates of the serial string out of the Motorola gps module. The Ref-0 just figures out what’s happening by watching those strings as they roll by. Hope that makes sense…. —————— If that’s all true, it should be “easy” to put dummy gps sentences into the Ref-0 interface port and a pps from something else. Once it sees stuff that it likes, it will start locking up to that pps. Cheap micro and just about any modern gps = fancy new GPSDO. Quick / easy / not much work. A little more complex if you decide to translate sawtooth data. That of course assumes these boxes do anything with sawtooth ... Bob > On Nov 6, 2014, at 7:52 AM, GandalfG8--- via time-nuts <time-nuts@febo.com> wrote: > > Hi Bob > > I view the master/slave situation the same way you do, but only  commented > because Lucent effectively sets things up the other way round so just  > wanted to be sure which way Paul was considering it. > > As regards, the "faking", yes, a single 15 way plug containing just  a link > and a resistor is exactly what I use now. > > I realise the RS422 fudge isn't ideal but certainly very handy for some  > quick tests. > > I've got two or three RS422 PCI interfaces but hadn't previously had any  > spare slots, now I've got a couple of lovely Magma 13 way PCI expansion units > but need to hack them about so the ports are at the front. > Whoever decided rack mount PC kit should be built with all the  connections > at the rear must have been really nuts, and it was too long ago  to use > "corridors" of racks as a viable excuse, someone somewhere just fitted  ears to > desk top style cases and left the rest of us to get on with  it.. > > One of the best things I ever did was to buy a pair of "back to front"  > rackmount PC cases with all ports and cards available at the front, trouble is  > now I want everything else to match, and most test gear has just the same  > issues! > > All good fun, and certainly never a dull moment:-) > > Regards > > Nigel > GM8PZR > > > _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
BC
Bob Camp
Fri, Nov 7, 2014 12:07 PM

Hi

When Satstat does the change, I’m sure it uses the correct command to do it. The syntax is a phrase followed by a numeric. You then use the same phrase followed by a question mark to see if it took it. The slave happy accepts the command, stores the data, and returns it when asked. A reasonable external program would indeed say “yup it did it”.

With no serial com (that’s a guess not a proven) there is no way to change the GPS. The slave simply stores the data. The only alternative would be for it to reject virtually all commands over the Diag port. Much easier to just put a note in the manual than to disable all that code.

Bob

On Nov 7, 2014, at 12:02 AM, Bob Stewart bob@evoria.net wrote:

Hi Bob,
I've been using the Satstat program to send commands.  What I don't understand is why the slave will accept commands such as sat elevation angle and not give an error response.  It gives a response of "command complete", but it doesn't change it.  Or is that an indication that both units use the same firmware, and success is actually indicated by the resulting change of the angle?

Bob    From: Bob Camp kb8tq@n1k.org
To: GandalfG8@aol.com; Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Thursday, November 6, 2014 6:07 PM
Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions

Hi

Ok, I got to spend a little quality time with my pair today. I mostly poked at the Diag port with a terminal program.

Both of the units seem talk over the Diag port regardless of the mode they are in (master / slave / fault ). Both respond to a sub-set of the Z3801 SCPI commands (= not all commands work). There are a few enhancements where a query to a “node” gives a reply (which it should not if it’s “node only”). All of it works ok at 9600 baud 8N1. All of it comes back with an error message (E-xxx) after a valid command response. It may not like the terminal program sending cr/lf or something like that.

By far the most useful thing to type is :SYST:STAT?
That executes the system:status request and gives you a nice full screen of information. It shows you about 95% of the useful information all on one page. About the only major thing missing is the DAC value. That’s available with another command out of the Z3801 list. It does take a while to compose and transmit the screen, so it’s not the best way to to this forever and ever.

The data from the request is the same on both boxes. Same everything except one shows as standby etc. The interesting details are things like elevation mask and cable delay. Why interesting? You can set them with discrete commands. You can also query the data with the same commands to make sure the box “took” what you sent.

This allows experimentation with changing the elevation mask on each box (or just simply doing a discrete query for the mask on each box). The Ref-0 / slave will happily change things around with discrete commands. None of them are reflected in the :SYST:STAT? listing on the Ref-0 or Ref-1. When you look at the Ref-1 / gps box, it’s discrete query information matches the :SYST:STAT? screen on both boxes. When you change things on the Ref-1 box both screens follow that data.

—————

When you pull the antenna on the gps box, it immediately (almost) goes into gps fault led flashing mode. The slave box happily sits there for a while (say a minute) with it’s gps led saying all is well. When you plug the antenna back in, the process is reversed. The gps box goes out of flash mode right away. The led goes out in a bit. The led on the slave box does not follow for a minute or so.

—————

All of this suggests to me that the Ref-0 / slave does not talk back to the other box via serial. It also suggests to me that all it “sees” from the Ref-1/gps box are the duplicates of the serial string out of the Motorola gps module. The Ref-0 just figures out what’s happening by watching those strings as they roll by.

Hope that makes sense….

——————

If that’s all true, it should be “easy” to put dummy gps sentences into the Ref-0 interface port and a pps from something else. Once it sees stuff that it likes, it will start locking up to that pps. Cheap micro and just about any modern gps = fancy new GPSDO. Quick / easy / not much work. A little more complex if you decide to translate sawtooth data. That of course assumes these boxes do anything with sawtooth ...

Bob

On Nov 6, 2014, at 7:52 AM, GandalfG8--- via time-nuts time-nuts@febo.com wrote:

Hi Bob

I view the master/slave situation the same way you do, but only  commented
because Lucent effectively sets things up the other way round so just
wanted to be sure which way Paul was considering it.

As regards, the "faking", yes, a single 15 way plug containing just  a link
and a resistor is exactly what I use now.

I realise the RS422 fudge isn't ideal but certainly very handy for some
quick tests.

I've got two or three RS422 PCI interfaces but hadn't previously had any
spare slots, now I've got a couple of lovely Magma 13 way PCI expansion units
but need to hack them about so the ports are at the front.
Whoever decided rack mount PC kit should be built with all the  connections
at the rear must have been really nuts, and it was too long ago  to use
"corridors" of racks as a viable excuse, someone somewhere just fitted  ears to
desk top style cases and left the rest of us to get on with  it..

One of the best things I ever did was to buy a pair of "back to front"
rackmount PC cases with all ports and cards available at the front, trouble is
now I want everything else to match, and most test gear has just the same
issues!

All good fun, and certainly never a dull moment:-)

Regards

Nigel
GM8PZR


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi When Satstat does the change, I’m sure it uses the correct command to do it. The syntax is a phrase followed by a numeric. You then use the same phrase followed by a question mark to see if it took it. The slave happy accepts the command, stores the data, and returns it when asked. A reasonable external program would indeed say “yup it did it”. With no serial com (that’s a guess not a proven) there is no way to change the GPS. The slave simply stores the data. The only alternative would be for it to reject virtually all commands over the Diag port. Much easier to just put a note in the manual than to disable all that code. Bob > On Nov 7, 2014, at 12:02 AM, Bob Stewart <bob@evoria.net> wrote: > > Hi Bob, > I've been using the Satstat program to send commands. What I don't understand is why the slave will accept commands such as sat elevation angle and not give an error response. It gives a response of "command complete", but it doesn't change it. Or is that an indication that both units use the same firmware, and success is actually indicated by the resulting change of the angle? > > Bob From: Bob Camp <kb8tq@n1k.org> > To: GandalfG8@aol.com; Discussion of precise time and frequency measurement <time-nuts@febo.com> > Sent: Thursday, November 6, 2014 6:07 PM > Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions > > Hi > > Ok, I got to spend a little quality time with my pair today. I mostly poked at the Diag port with a terminal program. > > Both of the units seem talk over the Diag port regardless of the mode they are in (master / slave / fault ). Both respond to a sub-set of the Z3801 SCPI commands (= not all commands work). There are a few enhancements where a query to a “node” gives a reply (which it should not if it’s “node only”). All of it works ok at 9600 baud 8N1. All of it comes back with an error message (E-xxx) after a valid command response. It may not like the terminal program sending cr/lf or something like that. > > By far the most useful thing to type is :SYST:STAT? > That executes the system:status request and gives you a nice full screen of information. It shows you about 95% of the useful information all on one page. About the only major thing missing is the DAC value. That’s available with another command out of the Z3801 list. It does take a while to compose and transmit the screen, so it’s not the best way to to this forever and ever. > > The data from the request is the same on both boxes. Same everything except one shows as standby etc. The interesting details are things like elevation mask and cable delay. Why interesting? You can set them with discrete commands. You can also query the data with the same commands to make sure the box “took” what you sent. > > This allows experimentation with changing the elevation mask on each box (or just simply doing a discrete query for the mask on each box). The Ref-0 / slave will happily change things around with discrete commands. None of them are reflected in the :SYST:STAT? listing on the Ref-0 or Ref-1. When you look at the Ref-1 / gps box, it’s discrete query information matches the :SYST:STAT? screen on both boxes. When you change things on the Ref-1 box both screens follow that data. > > ————— > > When you pull the antenna on the gps box, it immediately (almost) goes into gps fault led flashing mode. The slave box happily sits there for a while (say a minute) with it’s gps led saying all is well. When you plug the antenna back in, the process is reversed. The gps box goes out of flash mode right away. The led goes out in a bit. The led on the slave box does not follow for a minute or so. > > ————— > > All of this suggests to me that the Ref-0 / slave does not talk back to the other box via serial. It also suggests to me that all it “sees” from the Ref-1/gps box are the duplicates of the serial string out of the Motorola gps module. The Ref-0 just figures out what’s happening by watching those strings as they roll by. > > Hope that makes sense…. > > —————— > > If that’s all true, it should be “easy” to put dummy gps sentences into the Ref-0 interface port and a pps from something else. Once it sees stuff that it likes, it will start locking up to that pps. Cheap micro and just about any modern gps = fancy new GPSDO. Quick / easy / not much work. A little more complex if you decide to translate sawtooth data. That of course assumes these boxes do anything with sawtooth ... > > Bob > > > > > > >> On Nov 6, 2014, at 7:52 AM, GandalfG8--- via time-nuts <time-nuts@febo.com> wrote: >> >> Hi Bob >> >> I view the master/slave situation the same way you do, but only commented >> because Lucent effectively sets things up the other way round so just >> wanted to be sure which way Paul was considering it. >> >> As regards, the "faking", yes, a single 15 way plug containing just a link >> and a resistor is exactly what I use now. >> >> I realise the RS422 fudge isn't ideal but certainly very handy for some >> quick tests. >> >> I've got two or three RS422 PCI interfaces but hadn't previously had any >> spare slots, now I've got a couple of lovely Magma 13 way PCI expansion units >> but need to hack them about so the ports are at the front. >> Whoever decided rack mount PC kit should be built with all the connections >> at the rear must have been really nuts, and it was too long ago to use >> "corridors" of racks as a viable excuse, someone somewhere just fitted ears to >> desk top style cases and left the rest of us to get on with it.. >> >> One of the best things I ever did was to buy a pair of "back to front" >> rackmount PC cases with all ports and cards available at the front, trouble is >> now I want everything else to match, and most test gear has just the same >> issues! >> >> All good fun, and certainly never a dull moment:-) >> >> Regards >> >> Nigel >> GM8PZR >> >> >> > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BR
Bill Riches
Fri, Nov 7, 2014 1:04 PM

Hi Bob and group,

Where are you obtaining Satstat and what windoze version will it work on?  XP, W7 32 or 64 bit?  XP I guess?

Bill Riches
Cape May

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob Stewart
Sent: Friday, November 07, 2014 12:03 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions

Hi Bob,
I've been using the Satstat program to send commands.  What I don't understand is why the slave will accept commands such as sat elevation angle and not give an error response.  It gives a response of "command complete", but it doesn't change it.  Or is that an indication that both units use the same firmware, and success is actually indicated by the resulting change of the angle?

Bob    From: Bob Camp kb8tq@n1k.org
To: GandalfG8@aol.com; Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Thursday, November 6, 2014 6:07 PM
Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions


This email is free from viruses and malware because avast! Antivirus protection is active.
http://www.avast.com

Hi Bob and group, Where are you obtaining Satstat and what windoze version will it work on? XP, W7 32 or 64 bit? XP I guess? Bill Riches Cape May -----Original Message----- From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob Stewart Sent: Friday, November 07, 2014 12:03 AM To: Discussion of precise time and frequency measurement Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions Hi Bob, I've been using the Satstat program to send commands. What I don't understand is why the slave will accept commands such as sat elevation angle and not give an error response. It gives a response of "command complete", but it doesn't change it. Or is that an indication that both units use the same firmware, and success is actually indicated by the resulting change of the angle? Bob From: Bob Camp <kb8tq@n1k.org> To: GandalfG8@aol.com; Discussion of precise time and frequency measurement <time-nuts@febo.com> Sent: Thursday, November 6, 2014 6:07 PM Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions --- This email is free from viruses and malware because avast! Antivirus protection is active. http://www.avast.com
GR
Götz Romahn
Fri, Nov 7, 2014 3:42 PM

if you want to go the easy way to talk to REF-1 of REF-0 via the
J8-Diagnostic port try Ulrich Bangerts Z38XX software.
Here:
http://www.romahn.info/lucent/z3811.exe
you will find for download a patched and renamed version that directly
recognizes our nice new toys. After setting the right "Parameters" try
"View" or "Manual Command Entry".
cheers Götz

Am 07.11.2014 13:07, :

Hi

When Satstat does the change, I’m sure it uses the correct command to do it. The syntax is a phrase followed by a numeric. You then use the same phrase followed by a question mark to see if it took it. The slave happy accepts the command, stores the data, and returns it when asked. A reasonable external program would indeed say “yup it did it”.

With no serial com (that’s a guess not a proven) there is no way to change the GPS. The slave simply stores the data. The only alternative would be for it to reject virtually all commands over the Diag port. Much easier to just put a note in the manual than to disable all that code.

Bob

On Nov 7, 2014, at 12:02 AM, Bob Stewart bob@evoria.net wrote:

Hi Bob,
I've been using the Satstat program to send commands.  What I don't understand is why the slave will accept commands such as sat elevation angle and not give an error response.  It gives a response of "command complete", but it doesn't change it.  Or is that an indication that both units use the same firmware, and success is actually indicated by the resulting change of the angle?

Bob    From: Bob Camp kb8tq@n1k.org
To: GandalfG8@aol.com; Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Thursday, November 6, 2014 6:07 PM
Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions

Hi

Ok, I got to spend a little quality time with my pair today. I mostly poked at the Diag port with a terminal program.

Both of the units seem talk over the Diag port regardless of the mode they are in (master / slave / fault ). Both respond to a sub-set of the Z3801 SCPI commands (= not all commands work). There are a few enhancements where a query to a “node” gives a reply (which it should not if it’s “node only”). All of it works ok at 9600 baud 8N1. All of it comes back with an error message (E-xxx) after a valid command response. It may not like the terminal program sending cr/lf or something like that.

By far the most useful thing to type is :SYST:STAT?
That executes the system:status request and gives you a nice full screen of information. It shows you about 95% of the useful information all on one page. About the only major thing missing is the DAC value. That’s available with another command out of the Z3801 list. It does take a while to compose and transmit the screen, so it’s not the best way to to this forever and ever.

The data from the request is the same on both boxes. Same everything except one shows as standby etc. The interesting details are things like elevation mask and cable delay. Why interesting? You can set them with discrete commands. You can also query the data with the same commands to make sure the box “took” what you sent.

This allows experimentation with changing the elevation mask on each box (or just simply doing a discrete query for the mask on each box). The Ref-0 / slave will happily change things around with discrete commands. None of them are reflected in the :SYST:STAT? listing on the Ref-0 or Ref-1. When you look at the Ref-1 / gps box, it’s discrete query information matches the :SYST:STAT? screen on both boxes. When you change things on the Ref-1 box both screens follow that data.

—————

When you pull the antenna on the gps box, it immediately (almost) goes into gps fault led flashing mode. The slave box happily sits there for a while (say a minute) with it’s gps led saying all is well. When you plug the antenna back in, the process is reversed. The gps box goes out of flash mode right away. The led goes out in a bit. The led on the slave box does not follow for a minute or so.

—————

All of this suggests to me that the Ref-0 / slave does not talk back to the other box via serial. It also suggests to me that all it “sees” from the Ref-1/gps box are the duplicates of the serial string out of the Motorola gps module. The Ref-0 just figures out what’s happening by watching those strings as they roll by.

Hope that makes sense….

——————

If that’s all true, it should be “easy” to put dummy gps sentences into the Ref-0 interface port and a pps from something else. Once it sees stuff that it likes, it will start locking up to that pps. Cheap micro and just about any modern gps = fancy new GPSDO. Quick / easy / not much work. A little more complex if you decide to translate sawtooth data. That of course assumes these boxes do anything with sawtooth ...

Bob

On Nov 6, 2014, at 7:52 AM, GandalfG8--- via time-nuts time-nuts@febo.com wrote:

Hi Bob

I view the master/slave situation the same way you do, but only  commented
because Lucent effectively sets things up the other way round so just
wanted to be sure which way Paul was considering it.

As regards, the "faking", yes, a single 15 way plug containing just  a link
and a resistor is exactly what I use now.

I realise the RS422 fudge isn't ideal but certainly very handy for some
quick tests.

I've got two or three RS422 PCI interfaces but hadn't previously had any
spare slots, now I've got a couple of lovely Magma 13 way PCI expansion units
but need to hack them about so the ports are at the front.
Whoever decided rack mount PC kit should be built with all the  connections
at the rear must have been really nuts, and it was too long ago  to use
"corridors" of racks as a viable excuse, someone somewhere just fitted  ears to
desk top style cases and left the rest of us to get on with  it..

One of the best things I ever did was to buy a pair of "back to front"
rackmount PC cases with all ports and cards available at the front, trouble is
now I want everything else to match, and most test gear has just the same
issues!

All good fun, and certainly never a dull moment:-)

Regards

Nigel
GM8PZR


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

if you want to go the easy way to talk to REF-1 of REF-0 via the J8-Diagnostic port try Ulrich Bangerts Z38XX software. Here: http://www.romahn.info/lucent/z3811.exe you will find for download a patched and renamed version that directly recognizes our nice new toys. After setting the right "Parameters" try "View" or "Manual Command Entry". cheers Götz Am 07.11.2014 13:07, : > Hi > > When Satstat does the change, I’m sure it uses the correct command to do it. The syntax is a phrase followed by a numeric. You then use the same phrase followed by a question mark to see if it took it. The slave happy accepts the command, stores the data, and returns it when asked. A reasonable external program would indeed say “yup it did it”. > > With no serial com (that’s a guess not a proven) there is no way to change the GPS. The slave simply stores the data. The only alternative would be for it to reject virtually all commands over the Diag port. Much easier to just put a note in the manual than to disable all that code. > > Bob > >> On Nov 7, 2014, at 12:02 AM, Bob Stewart <bob@evoria.net> wrote: >> >> Hi Bob, >> I've been using the Satstat program to send commands. What I don't understand is why the slave will accept commands such as sat elevation angle and not give an error response. It gives a response of "command complete", but it doesn't change it. Or is that an indication that both units use the same firmware, and success is actually indicated by the resulting change of the angle? >> >> Bob From: Bob Camp <kb8tq@n1k.org> >> To: GandalfG8@aol.com; Discussion of precise time and frequency measurement <time-nuts@febo.com> >> Sent: Thursday, November 6, 2014 6:07 PM >> Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions >> >> Hi >> >> Ok, I got to spend a little quality time with my pair today. I mostly poked at the Diag port with a terminal program. >> >> Both of the units seem talk over the Diag port regardless of the mode they are in (master / slave / fault ). Both respond to a sub-set of the Z3801 SCPI commands (= not all commands work). There are a few enhancements where a query to a “node” gives a reply (which it should not if it’s “node only”). All of it works ok at 9600 baud 8N1. All of it comes back with an error message (E-xxx) after a valid command response. It may not like the terminal program sending cr/lf or something like that. >> >> By far the most useful thing to type is :SYST:STAT? >> That executes the system:status request and gives you a nice full screen of information. It shows you about 95% of the useful information all on one page. About the only major thing missing is the DAC value. That’s available with another command out of the Z3801 list. It does take a while to compose and transmit the screen, so it’s not the best way to to this forever and ever. >> >> The data from the request is the same on both boxes. Same everything except one shows as standby etc. The interesting details are things like elevation mask and cable delay. Why interesting? You can set them with discrete commands. You can also query the data with the same commands to make sure the box “took” what you sent. >> >> This allows experimentation with changing the elevation mask on each box (or just simply doing a discrete query for the mask on each box). The Ref-0 / slave will happily change things around with discrete commands. None of them are reflected in the :SYST:STAT? listing on the Ref-0 or Ref-1. When you look at the Ref-1 / gps box, it’s discrete query information matches the :SYST:STAT? screen on both boxes. When you change things on the Ref-1 box both screens follow that data. >> >> ————— >> >> When you pull the antenna on the gps box, it immediately (almost) goes into gps fault led flashing mode. The slave box happily sits there for a while (say a minute) with it’s gps led saying all is well. When you plug the antenna back in, the process is reversed. The gps box goes out of flash mode right away. The led goes out in a bit. The led on the slave box does not follow for a minute or so. >> >> ————— >> >> All of this suggests to me that the Ref-0 / slave does not talk back to the other box via serial. It also suggests to me that all it “sees” from the Ref-1/gps box are the duplicates of the serial string out of the Motorola gps module. The Ref-0 just figures out what’s happening by watching those strings as they roll by. >> >> Hope that makes sense…. >> >> —————— >> >> If that’s all true, it should be “easy” to put dummy gps sentences into the Ref-0 interface port and a pps from something else. Once it sees stuff that it likes, it will start locking up to that pps. Cheap micro and just about any modern gps = fancy new GPSDO. Quick / easy / not much work. A little more complex if you decide to translate sawtooth data. That of course assumes these boxes do anything with sawtooth ... >> >> Bob >> >> >> >> >> >> >>> On Nov 6, 2014, at 7:52 AM, GandalfG8--- via time-nuts <time-nuts@febo.com> wrote: >>> >>> Hi Bob >>> >>> I view the master/slave situation the same way you do, but only commented >>> because Lucent effectively sets things up the other way round so just >>> wanted to be sure which way Paul was considering it. >>> >>> As regards, the "faking", yes, a single 15 way plug containing just a link >>> and a resistor is exactly what I use now. >>> >>> I realise the RS422 fudge isn't ideal but certainly very handy for some >>> quick tests. >>> >>> I've got two or three RS422 PCI interfaces but hadn't previously had any >>> spare slots, now I've got a couple of lovely Magma 13 way PCI expansion units >>> but need to hack them about so the ports are at the front. >>> Whoever decided rack mount PC kit should be built with all the connections >>> at the rear must have been really nuts, and it was too long ago to use >>> "corridors" of racks as a viable excuse, someone somewhere just fitted ears to >>> desk top style cases and left the rest of us to get on with it.. >>> >>> One of the best things I ever did was to buy a pair of "back to front" >>> rackmount PC cases with all ports and cards available at the front, trouble is >>> now I want everything else to match, and most test gear has just the same >>> issues! >>> >>> All good fun, and certainly never a dull moment:-) >>> >>> Regards >>> >>> Nigel >>> GM8PZR >>> >>> >>> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
BS
Bob Stewart
Fri, Nov 7, 2014 6:34 PM

Hi Bill,

I think I got my copy here.  I used the bottom one.  I don't know what a 58503 is.  it was hard to find, for some reason.  I'm using XP on an old laptop.

http://www.ko4bb.com/manuals/index.php?dir=05%29_GPS_Timing/SatStat
Bob
From: Bill Riches bill.riches@verizon.net
To: 'Bob Stewart' bob@evoria.net; 'Discussion of precise time and frequency measurement' time-nuts@febo.com
Sent: Friday, November 7, 2014 7:04 AM
Subject: RE: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions

Hi Bob and group,

Where are you obtaining Satstat and what windoze version will it work on?  XP, W7 32 or 64 bit?  XP I guess?

Bill Riches
Cape May

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob Stewart
Sent: Friday, November 07, 2014 12:03 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions

Hi Bob,
I've been using the Satstat program to send commands.  What I don't understand is why the slave will accept commands such as sat elevation angle and not give an error response.  It gives a response of "command complete", but it doesn't change it.  Or is that an indication that both units use the same firmware, and success is actually indicated by the resulting change of the angle?

Bob    From: Bob Camp kb8tq@n1k.org
To: GandalfG8@aol.com; Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Thursday, November 6, 2014 6:07 PM
Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions


This email is free from viruses and malware because avast! Antivirus protection is active.
http://www.avast.com

Hi Bill, I think I got my copy here.  I used the bottom one.  I don't know what a 58503 is.  it was hard to find, for some reason.  I'm using XP on an old laptop. http://www.ko4bb.com/manuals/index.php?dir=05%29_GPS_Timing/SatStat Bob From: Bill Riches <bill.riches@verizon.net> To: 'Bob Stewart' <bob@evoria.net>; 'Discussion of precise time and frequency measurement' <time-nuts@febo.com> Sent: Friday, November 7, 2014 7:04 AM Subject: RE: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions Hi Bob and group, Where are you obtaining Satstat and what windoze version will it work on?  XP, W7 32 or 64 bit?  XP I guess? Bill Riches Cape May -----Original Message----- From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob Stewart Sent: Friday, November 07, 2014 12:03 AM To: Discussion of precise time and frequency measurement Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions Hi Bob, I've been using the Satstat program to send commands.  What I don't understand is why the slave will accept commands such as sat elevation angle and not give an error response.  It gives a response of "command complete", but it doesn't change it.  Or is that an indication that both units use the same firmware, and success is actually indicated by the resulting change of the angle? Bob    From: Bob Camp <kb8tq@n1k.org> To: GandalfG8@aol.com; Discussion of precise time and frequency measurement <time-nuts@febo.com> Sent: Thursday, November 6, 2014 6:07 PM Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions --- This email is free from viruses and malware because avast! Antivirus protection is active. http://www.avast.com
X
xaos
Fri, Nov 7, 2014 8:41 PM

It seems that this program will not work for Win 7.
Has anyone attempted to decompile this thing?

Maybe re-work a bit and have it run on anything?
-George, N2FGX

On 11/07/2014 01:34 PM, Bob Stewart wrote:

Hi Bill,

I think I got my copy here.  I used the bottom one.  I don't know what a 58503 is.  it was hard to find, for some reason.  I'm using XP on an old laptop.

http://www.ko4bb.com/manuals/index.php?dir=05%29_GPS_Timing/SatStat
Bob
From: Bill Riches bill.riches@verizon.net
To: 'Bob Stewart' bob@evoria.net; 'Discussion of precise time and frequency measurement' time-nuts@febo.com
Sent: Friday, November 7, 2014 7:04 AM
Subject: RE: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions

Hi Bob and group,

Where are you obtaining Satstat and what windoze version will it work on?  XP, W7 32 or 64 bit?  XP I guess?

Bill Riches
Cape May

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob Stewart
Sent: Friday, November 07, 2014 12:03 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions

Hi Bob,
I've been using the Satstat program to send commands.  What I don't understand is why the slave will accept commands such as sat elevation angle and not give an error response.  It gives a response of "command complete", but it doesn't change it.  Or is that an indication that both units use the same firmware, and success is actually indicated by the resulting change of the angle?

Bob    From: Bob Camp kb8tq@n1k.org
To: GandalfG8@aol.com; Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Thursday, November 6, 2014 6:07 PM
Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions


This email is free from viruses and malware because avast! Antivirus protection is active.
http://www.avast.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

It seems that this program will not work for Win 7. Has anyone attempted to decompile this thing? Maybe re-work a bit and have it run on anything? -George, N2FGX On 11/07/2014 01:34 PM, Bob Stewart wrote: > Hi Bill, > > I think I got my copy here. I used the bottom one. I don't know what a 58503 is. it was hard to find, for some reason. I'm using XP on an old laptop. > > http://www.ko4bb.com/manuals/index.php?dir=05%29_GPS_Timing/SatStat > Bob > From: Bill Riches <bill.riches@verizon.net> > To: 'Bob Stewart' <bob@evoria.net>; 'Discussion of precise time and frequency measurement' <time-nuts@febo.com> > Sent: Friday, November 7, 2014 7:04 AM > Subject: RE: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions > > Hi Bob and group, > > Where are you obtaining Satstat and what windoze version will it work on? XP, W7 32 or 64 bit? XP I guess? > > Bill Riches > Cape May > > > > -----Original Message----- > From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob Stewart > Sent: Friday, November 07, 2014 12:03 AM > To: Discussion of precise time and frequency measurement > Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions > > Hi Bob, > I've been using the Satstat program to send commands. What I don't understand is why the slave will accept commands such as sat elevation angle and not give an error response. It gives a response of "command complete", but it doesn't change it. Or is that an indication that both units use the same firmware, and success is actually indicated by the resulting change of the angle? > > Bob From: Bob Camp <kb8tq@n1k.org> > To: GandalfG8@aol.com; Discussion of precise time and frequency measurement <time-nuts@febo.com> > Sent: Thursday, November 6, 2014 6:07 PM > Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions > > > --- > This email is free from viruses and malware because avast! Antivirus protection is active. > http://www.avast.com > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BC
Bob Camp
Sat, Nov 8, 2014 1:16 AM

Hi

Thanks for posting that information. Keeping Ulrich’s code going is a very good thing.

Bob

On Nov 7, 2014, at 10:42 AM, Götz Romahn goetz@g-romahn.de wrote:

if you want to go the easy way to talk to REF-1 of REF-0 via the J8-Diagnostic port try Ulrich Bangerts Z38XX software.
Here:
http://www.romahn.info/lucent/z3811.exe
you will find for download a patched and renamed version that directly recognizes our nice new toys. After setting the right "Parameters" try "View" or "Manual Command Entry".
cheers Götz

Am 07.11.2014 13:07, :

Hi

When Satstat does the change, I’m sure it uses the correct command to do it. The syntax is a phrase followed by a numeric. You then use the same phrase followed by a question mark to see if it took it. The slave happy accepts the command, stores the data, and returns it when asked. A reasonable external program would indeed say “yup it did it”.

With no serial com (that’s a guess not a proven) there is no way to change the GPS. The slave simply stores the data. The only alternative would be for it to reject virtually all commands over the Diag port. Much easier to just put a note in the manual than to disable all that code.

Bob

On Nov 7, 2014, at 12:02 AM, Bob Stewart bob@evoria.net wrote:

Hi Bob,
I've been using the Satstat program to send commands.  What I don't understand is why the slave will accept commands such as sat elevation angle and not give an error response.  It gives a response of "command complete", but it doesn't change it.  Or is that an indication that both units use the same firmware, and success is actually indicated by the resulting change of the angle?

Bob    From: Bob Camp kb8tq@n1k.org
To: GandalfG8@aol.com; Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Thursday, November 6, 2014 6:07 PM
Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions

Hi

Ok, I got to spend a little quality time with my pair today. I mostly poked at the Diag port with a terminal program.

Both of the units seem talk over the Diag port regardless of the mode they are in (master / slave / fault ). Both respond to a sub-set of the Z3801 SCPI commands (= not all commands work). There are a few enhancements where a query to a “node” gives a reply (which it should not if it’s “node only”). All of it works ok at 9600 baud 8N1. All of it comes back with an error message (E-xxx) after a valid command response. It may not like the terminal program sending cr/lf or something like that.

By far the most useful thing to type is :SYST:STAT?
That executes the system:status request and gives you a nice full screen of information. It shows you about 95% of the useful information all on one page. About the only major thing missing is the DAC value. That’s available with another command out of the Z3801 list. It does take a while to compose and transmit the screen, so it’s not the best way to to this forever and ever.

The data from the request is the same on both boxes. Same everything except one shows as standby etc. The interesting details are things like elevation mask and cable delay. Why interesting? You can set them with discrete commands. You can also query the data with the same commands to make sure the box “took” what you sent.

This allows experimentation with changing the elevation mask on each box (or just simply doing a discrete query for the mask on each box). The Ref-0 / slave will happily change things around with discrete commands. None of them are reflected in the :SYST:STAT? listing on the Ref-0 or Ref-1. When you look at the Ref-1 / gps box, it’s discrete query information matches the :SYST:STAT? screen on both boxes. When you change things on the Ref-1 box both screens follow that data.

—————

When you pull the antenna on the gps box, it immediately (almost) goes into gps fault led flashing mode. The slave box happily sits there for a while (say a minute) with it’s gps led saying all is well. When you plug the antenna back in, the process is reversed. The gps box goes out of flash mode right away. The led goes out in a bit. The led on the slave box does not follow for a minute or so.

—————

All of this suggests to me that the Ref-0 / slave does not talk back to the other box via serial. It also suggests to me that all it “sees” from the Ref-1/gps box are the duplicates of the serial string out of the Motorola gps module. The Ref-0 just figures out what’s happening by watching those strings as they roll by.

Hope that makes sense….

——————

If that’s all true, it should be “easy” to put dummy gps sentences into the Ref-0 interface port and a pps from something else. Once it sees stuff that it likes, it will start locking up to that pps. Cheap micro and just about any modern gps = fancy new GPSDO. Quick / easy / not much work. A little more complex if you decide to translate sawtooth data. That of course assumes these boxes do anything with sawtooth ...

Bob

On Nov 6, 2014, at 7:52 AM, GandalfG8--- via time-nuts time-nuts@febo.com wrote:

Hi Bob

I view the master/slave situation the same way you do, but only  commented
because Lucent effectively sets things up the other way round so just
wanted to be sure which way Paul was considering it.

As regards, the "faking", yes, a single 15 way plug containing just  a link
and a resistor is exactly what I use now.

I realise the RS422 fudge isn't ideal but certainly very handy for some
quick tests.

I've got two or three RS422 PCI interfaces but hadn't previously had any
spare slots, now I've got a couple of lovely Magma 13 way PCI expansion units
but need to hack them about so the ports are at the front.
Whoever decided rack mount PC kit should be built with all the  connections
at the rear must have been really nuts, and it was too long ago  to use
"corridors" of racks as a viable excuse, someone somewhere just fitted  ears to
desk top style cases and left the rest of us to get on with  it..

One of the best things I ever did was to buy a pair of "back to front"
rackmount PC cases with all ports and cards available at the front, trouble is
now I want everything else to match, and most test gear has just the same
issues!

All good fun, and certainly never a dull moment:-)

Regards

Nigel
GM8PZR


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Thanks for posting that information. Keeping Ulrich’s code going is a very good thing. Bob > On Nov 7, 2014, at 10:42 AM, Götz Romahn <goetz@g-romahn.de> wrote: > > if you want to go the easy way to talk to REF-1 of REF-0 via the J8-Diagnostic port try Ulrich Bangerts Z38XX software. > Here: > http://www.romahn.info/lucent/z3811.exe > you will find for download a patched and renamed version that directly recognizes our nice new toys. After setting the right "Parameters" try "View" or "Manual Command Entry". > cheers Götz > > > Am 07.11.2014 13:07, : >> Hi >> >> When Satstat does the change, I’m sure it uses the correct command to do it. The syntax is a phrase followed by a numeric. You then use the same phrase followed by a question mark to see if it took it. The slave happy accepts the command, stores the data, and returns it when asked. A reasonable external program would indeed say “yup it did it”. >> >> With no serial com (that’s a guess not a proven) there is no way to change the GPS. The slave simply stores the data. The only alternative would be for it to reject virtually all commands over the Diag port. Much easier to just put a note in the manual than to disable all that code. >> >> Bob >> >>> On Nov 7, 2014, at 12:02 AM, Bob Stewart <bob@evoria.net> wrote: >>> >>> Hi Bob, >>> I've been using the Satstat program to send commands. What I don't understand is why the slave will accept commands such as sat elevation angle and not give an error response. It gives a response of "command complete", but it doesn't change it. Or is that an indication that both units use the same firmware, and success is actually indicated by the resulting change of the angle? >>> >>> Bob From: Bob Camp <kb8tq@n1k.org> >>> To: GandalfG8@aol.com; Discussion of precise time and frequency measurement <time-nuts@febo.com> >>> Sent: Thursday, November 6, 2014 6:07 PM >>> Subject: Re: [time-nuts] Lucent KS-24361 GPSDO arrived today several questions >>> >>> Hi >>> >>> Ok, I got to spend a little quality time with my pair today. I mostly poked at the Diag port with a terminal program. >>> >>> Both of the units seem talk over the Diag port regardless of the mode they are in (master / slave / fault ). Both respond to a sub-set of the Z3801 SCPI commands (= not all commands work). There are a few enhancements where a query to a “node” gives a reply (which it should not if it’s “node only”). All of it works ok at 9600 baud 8N1. All of it comes back with an error message (E-xxx) after a valid command response. It may not like the terminal program sending cr/lf or something like that. >>> >>> By far the most useful thing to type is :SYST:STAT? >>> That executes the system:status request and gives you a nice full screen of information. It shows you about 95% of the useful information all on one page. About the only major thing missing is the DAC value. That’s available with another command out of the Z3801 list. It does take a while to compose and transmit the screen, so it’s not the best way to to this forever and ever. >>> >>> The data from the request is the same on both boxes. Same everything except one shows as standby etc. The interesting details are things like elevation mask and cable delay. Why interesting? You can set them with discrete commands. You can also query the data with the same commands to make sure the box “took” what you sent. >>> >>> This allows experimentation with changing the elevation mask on each box (or just simply doing a discrete query for the mask on each box). The Ref-0 / slave will happily change things around with discrete commands. None of them are reflected in the :SYST:STAT? listing on the Ref-0 or Ref-1. When you look at the Ref-1 / gps box, it’s discrete query information matches the :SYST:STAT? screen on both boxes. When you change things on the Ref-1 box both screens follow that data. >>> >>> ————— >>> >>> When you pull the antenna on the gps box, it immediately (almost) goes into gps fault led flashing mode. The slave box happily sits there for a while (say a minute) with it’s gps led saying all is well. When you plug the antenna back in, the process is reversed. The gps box goes out of flash mode right away. The led goes out in a bit. The led on the slave box does not follow for a minute or so. >>> >>> ————— >>> >>> All of this suggests to me that the Ref-0 / slave does not talk back to the other box via serial. It also suggests to me that all it “sees” from the Ref-1/gps box are the duplicates of the serial string out of the Motorola gps module. The Ref-0 just figures out what’s happening by watching those strings as they roll by. >>> >>> Hope that makes sense…. >>> >>> —————— >>> >>> If that’s all true, it should be “easy” to put dummy gps sentences into the Ref-0 interface port and a pps from something else. Once it sees stuff that it likes, it will start locking up to that pps. Cheap micro and just about any modern gps = fancy new GPSDO. Quick / easy / not much work. A little more complex if you decide to translate sawtooth data. That of course assumes these boxes do anything with sawtooth ... >>> >>> Bob >>> >>> >>> >>> >>> >>> >>>> On Nov 6, 2014, at 7:52 AM, GandalfG8--- via time-nuts <time-nuts@febo.com> wrote: >>>> >>>> Hi Bob >>>> >>>> I view the master/slave situation the same way you do, but only commented >>>> because Lucent effectively sets things up the other way round so just >>>> wanted to be sure which way Paul was considering it. >>>> >>>> As regards, the "faking", yes, a single 15 way plug containing just a link >>>> and a resistor is exactly what I use now. >>>> >>>> I realise the RS422 fudge isn't ideal but certainly very handy for some >>>> quick tests. >>>> >>>> I've got two or three RS422 PCI interfaces but hadn't previously had any >>>> spare slots, now I've got a couple of lovely Magma 13 way PCI expansion units >>>> but need to hack them about so the ports are at the front. >>>> Whoever decided rack mount PC kit should be built with all the connections >>>> at the rear must have been really nuts, and it was too long ago to use >>>> "corridors" of racks as a viable excuse, someone somewhere just fitted ears to >>>> desk top style cases and left the rest of us to get on with it.. >>>> >>>> One of the best things I ever did was to buy a pair of "back to front" >>>> rackmount PC cases with all ports and cards available at the front, trouble is >>>> now I want everything else to match, and most test gear has just the same >>>> issues! >>>> >>>> All good fun, and certainly never a dull moment:-) >>>> >>>> Regards >>>> >>>> Nigel >>>> GM8PZR >>>> >>>> >>>> >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >>> >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
AH
Alan Hochhalter
Sat, Nov 8, 2014 2:09 AM

I got in a hurry and made mistake #1 listed by Bob a few replies back.  I
hooked one up to the RS422/1PPI port transmit pins (same as on J8) and was
receiving data fine in the PC so I just hooked up the other pair of
terminals to the receive pins for J8 and couldn't get any response to
commands.

I finally found that I had initially connected the Lucent box ouput pins to
the wrong terminals on the converter.  The Time code data on J6 was still
getting into the PC even though it was connected wrong.  So I just wired
the Lucent box input pins to the other pins on the converter that are
really inputs and couldn't get J8 sto work either.

What I ended up with to make it work is: J8-9 to RX-, J8-5 to RX+, J8-8 to
TX- and T8-4 to TX+.

Alan

I got in a hurry and made mistake #1 listed by Bob a few replies back. I hooked one up to the RS422/1PPI port transmit pins (same as on J8) and was receiving data fine in the PC so I just hooked up the other pair of terminals to the receive pins for J8 and couldn't get any response to commands. I finally found that I had initially connected the Lucent box ouput pins to the wrong terminals on the converter. The Time code data on J6 was still getting into the PC even though it was connected wrong. So I just wired the Lucent box input pins to the other pins on the converter that are really inputs and couldn't get J8 sto work either. What I ended up with to make it work is: J8-9 to RX-, J8-5 to RX+, J8-8 to TX- and T8-4 to TX+. Alan