Thanks Demetrios, that was a very interesting read.
And thanks everyone else for responding.
Cheers
Michael
On Sat, 18 Mar 2023 at 3:02 am, Demetrios Matsakis via time-nuts <
time-nuts@lists.febo.com> wrote:
I don’t know how SR counters are today, but when we were upgrading our
infrastructure over a decade ago we found other counters had better
linearity. Rover et al’s open source article has a good discussion of
these issues, although of course you need to have one if you are going to
experiment. See G. D. Rovera, M. Siccardi, S. Romisch, and M. Abgrali,
“Time delay measurements: estimation of the error budget”, Metrologia 56,
2019 035004
On Mar 17, 2023, at 9:46 AM, Magnus Danielson via time-nuts <
time-nuts@lists.febo.com> wrote:
Dear Michael,
On 2023-03-16 08:17, Michael Wouters via time-nuts wrote:
Dear time-nuts
Counter specs often include an “internal resolution” error. For example,
the SR620 specs say that it is 25 ps in single-shot, but this can be
reduced to 4 ps with sufficient, repeated measurements. Can anyone
offer
any enlightenment as to the origin of this error, and the statistical
distribution it has? I mentioned the SR620 but information about the
53230A
would be interesting too.
First of all, the single-shot resolution is somewhat of a hallmark
measure when it comes to counters.
The interpolator resolution is part of this, but consider that there
exists non-linearities in the interpolator which makes the error larger. I
recall there being a plot of the non-linearity in the SR620 manual.
It is not uncommon to have interpolator resolution better than
non-linearities, but the later may be more subtle to most.
Averaging can help, but depending to specifics, it's hard to give a
number.
Cheers,
Magnus
Cheers
Michael
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
On Thu, 16 Mar 2023 at 19:06, Michael Wouters via time-nuts <
time-nuts@lists.febo.com> wrote:
Dear time-nuts
Counter specs often include an “internal resolution” error. For example,
the SR620 specs say that it is 25 ps in single-shot, but this can be
reduced to 4 ps with sufficient, repeated measurements. Can anyone offer
any enlightenment as to the origin of this error, and the statistical
distribution it has? I mentioned the SR620 but information about the 53230A
would be interesting too.
Cheers
Michael
I would have thought that if there were no systematic errors you could
reduce the average error to zero with an infinite amount of samples. But a
systematic error would prevent that. I don’t have an SR620, but it must
have a maximum amount of averaging.
I’m not a mathematician, but I if I had to take a guess, I would have
thought that the average error would reduce as the square root of the
number is samples. But (25/4)^2 is only 39, which is a lot less than the
maximum averaging the instrument can do. It would be interesting to either
find a formula, or even verify with a numerical simulation if I’m right
about the square root relationship.
Dr. David Kirkby,
Kirkby Microwave Ltd,
drkirkby@kirkbymicrowave.co.uk
https://www.kirkbymicrowave.co.uk/
Telephone 01621-680100./ +44 1621 680100
Registered in England & Wales, company number 08914892.
Registered office:
Stokes Hall Lodge, Burnham Rd, Althorne, Chelmsford, Essex, CM3 6DT, United
Kingdom
Hi
The random errors have a standard deviation. The “one sigma” is often what is
used to characterize them rather than the average. If they spec 10 ps resolution
and that’s limited by the random stuff, they are talking about 10 ps one sigma .
Bob
On Mar 19, 2023, at 9:27 AM, Dr. David Kirkby via time-nuts time-nuts@lists.febo.com wrote:
On Thu, 16 Mar 2023 at 19:06, Michael Wouters via time-nuts <
time-nuts@lists.febo.com> wrote:
Dear time-nuts
Counter specs often include an “internal resolution” error. For example,
the SR620 specs say that it is 25 ps in single-shot, but this can be
reduced to 4 ps with sufficient, repeated measurements. Can anyone offer
any enlightenment as to the origin of this error, and the statistical
distribution it has? I mentioned the SR620 but information about the 53230A
would be interesting too.
Cheers
Michael
I would have thought that if there were no systematic errors you could
reduce the average error to zero with an infinite amount of samples. But a
systematic error would prevent that. I don’t have an SR620, but it must
have a maximum amount of averaging.
I’m not a mathematician, but I if I had to take a guess, I would have
thought that the average error would reduce as the square root of the
number is samples. But (25/4)^2 is only 39, which is a lot less than the
maximum averaging the instrument can do. It would be interesting to either
find a formula, or even verify with a numerical simulation if I’m right
about the square root relationship.
Dr. David Kirkby,
Kirkby Microwave Ltd,
drkirkby@kirkbymicrowave.co.uk
https://www.kirkbymicrowave.co.uk/
Telephone 01621-680100./ +44 1621 680100
Registered in England & Wales, company number 08914892.
Registered office:
Stokes Hall Lodge, Burnham Rd, Althorne, Chelmsford, Essex, CM3 6DT, United
Kingdom
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com