Hi all,
I've started to see if is possible to fix an H-maser that we have here in
our timelab. Today we're contributing to UTC With some ensemble that is
composed by: 1 commercial cs clock + BVA phase locked in long term (<100s)
to the cs + synthesizer + HROG-5. And these are results for last circular
t: Cirt 460 LRTE PPP link
https://webtai.bipm.org/ftp/pub/tai/timelinks/lkc/2604/lrteptb/lnk/lrteptb.333a_.gif
I
wish to know if this result is good for an ensemble like we have here for
starting this discussion... And what you all suggest for improvement. Today
i have:
Real time tropospheric delay calculation with hopfield's model to replace
"MDTR" (with also mapping function to correct for each satellite). Every 15
minutes, also, I fed the Rx5Tr receiver with temperature, humidity and
pressure for its tropospheric model update. The sensor is located near the
antenna.
And an algorithm that simply sees the frequency deviation every 2 hours and
computes a frequency correction for our HROG-5.
We also have a hydrogen maser that is stopped because the 10 MHz oscillator
is not locking. I want to do some maintenance there to see if it can be
brought back to life.
The maser is glowing purple, but it shows "no vacuum" and "synthesizer is
faulty". I can see an A.M. oscillation of about -140 to -120 dBm in the
cavity output, but nothing more.
So, if possible, I may want to be guided in some steps to see if is it
possible to recover it.
Best regards,
Luiz
Hi all,
I've started to see if is possible to fix an H-maser that we have here in
our timelab. Today we're contributing to UTC With some ensemble that is
composed by: 1 commercial cs clock + BVA phase locked in long term (<100s)
to the cs + synthesizer + HROG-5. And these are results for last circular
t: Cirt 460 LRTE PPP link
<https://webtai.bipm.org/ftp/pub/tai/timelinks/lkc/2604/lrteptb/lnk/lrteptb.333a_.gif>
I
wish to know if this result is good for an ensemble like we have here for
starting this discussion... And what you all suggest for improvement. Today
i have:
Real time tropospheric delay calculation with hopfield's model to replace
"MDTR" (with also mapping function to correct for each satellite). Every 15
minutes, also, I fed the Rx5Tr receiver with temperature, humidity and
pressure for its tropospheric model update. The sensor is located near the
antenna.
And an algorithm that simply sees the frequency deviation every 2 hours and
computes a frequency correction for our HROG-5.
We also have a hydrogen maser that is stopped because the 10 MHz oscillator
is not locking. I want to do some maintenance there to see if it can be
brought back to life.
The maser is glowing purple, but it shows "no vacuum" and "synthesizer is
faulty". I can see an A.M. oscillation of about -140 to -120 dBm in the
cavity output, but nothing more.
So, if possible, I may want to be guided in some steps to see if is it
possible to recover it.
Best regards,
Luiz
J
john@miles.io
Tue, May 26, 2026 5:48 AM
The Cs+GNSS ensemble looks like it's performing very well. That's the
Septentrio PolaRx5TR receiver, correct?
It sounds like you may have a power supply problem with the maser, or
perhaps a controller issue, since you obviously have vacuum if the
dissociator is lit. How far off frequency is it? More than a few Hz
suggests that the oscillator oven has failed, which could explain everything
except the vacuum error.
-- john
-----Original Message-----
From: Luiz Paulo Damaceno via time-nuts time-nuts@lists.febo.com
Sent: Monday, May 25, 2026 11:39 AM
To: Discussion of precise time and frequency measurement
time-nuts@lists.febo.com
Cc: Luiz Paulo Damaceno luizpauloeletrico42@gmail.com
Subject: [time-nuts] UTC(LRTE) H-Maser maintenace
Hi all,
I've started to see if is possible to fix an H-maser that we have here in
our timelab. Today we're contributing to UTC With some ensemble that is
composed by: 1 commercial cs clock + BVA phase locked in long term (<100s)
to the cs + synthesizer + HROG-5. And these are results for last circular
t: Cirt 460 LRTE PPP link
<https://webtai.bipm.org/ftp/pub/tai/timelinks/lkc/2604/lrteptb/lnk/lrteptb.
333a_.gif>
I
wish to know if this result is good for an ensemble like we have here for
starting this discussion... And what you all suggest for improvement. Today
i have:
Real time tropospheric delay calculation with hopfield's model to replace
"MDTR" (with also mapping function to correct for each satellite). Every 15
minutes, also, I fed the Rx5Tr receiver with temperature, humidity and
pressure for its tropospheric model update. The sensor is located near the
antenna.
And an algorithm that simply sees the frequency deviation every 2 hours and
computes a frequency correction for our HROG-5.
We also have a hydrogen maser that is stopped because the 10 MHz oscillator
is not locking. I want to do some maintenance there to see if it can be
brought back to life.
The maser is glowing purple, but it shows "no vacuum" and "synthesizer is
faulty". I can see an A.M. oscillation of about -140 to -120 dBm in the
cavity output, but nothing more.
So, if possible, I may want to be guided in some steps to see if is it
possible to recover it.
Best regards,
Luiz
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe send an
email to time-nuts-leave@lists.febo.com
The Cs+GNSS ensemble looks like it's performing very well. That's the
Septentrio PolaRx5TR receiver, correct?
It sounds like you may have a power supply problem with the maser, or
perhaps a controller issue, since you obviously have vacuum if the
dissociator is lit. How far off frequency is it? More than a few Hz
suggests that the oscillator oven has failed, which could explain everything
except the vacuum error.
-- john
-----Original Message-----
From: Luiz Paulo Damaceno via time-nuts <time-nuts@lists.febo.com>
Sent: Monday, May 25, 2026 11:39 AM
To: Discussion of precise time and frequency measurement
<time-nuts@lists.febo.com>
Cc: Luiz Paulo Damaceno <luizpauloeletrico42@gmail.com>
Subject: [time-nuts] UTC(LRTE) H-Maser maintenace
Hi all,
I've started to see if is possible to fix an H-maser that we have here in
our timelab. Today we're contributing to UTC With some ensemble that is
composed by: 1 commercial cs clock + BVA phase locked in long term (<100s)
to the cs + synthesizer + HROG-5. And these are results for last circular
t: Cirt 460 LRTE PPP link
<https://webtai.bipm.org/ftp/pub/tai/timelinks/lkc/2604/lrteptb/lnk/lrteptb.
333a_.gif>
I
wish to know if this result is good for an ensemble like we have here for
starting this discussion... And what you all suggest for improvement. Today
i have:
Real time tropospheric delay calculation with hopfield's model to replace
"MDTR" (with also mapping function to correct for each satellite). Every 15
minutes, also, I fed the Rx5Tr receiver with temperature, humidity and
pressure for its tropospheric model update. The sensor is located near the
antenna.
And an algorithm that simply sees the frequency deviation every 2 hours and
computes a frequency correction for our HROG-5.
We also have a hydrogen maser that is stopped because the 10 MHz oscillator
is not locking. I want to do some maintenance there to see if it can be
brought back to life.
The maser is glowing purple, but it shows "no vacuum" and "synthesizer is
faulty". I can see an A.M. oscillation of about -140 to -120 dBm in the
cavity output, but nothing more.
So, if possible, I may want to be guided in some steps to see if is it
possible to recover it.
Best regards,
Luiz
_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe send an
email to time-nuts-leave@lists.febo.com
AK
Attila Kinali
Tue, May 26, 2026 6:04 AM
A wonderful good morning!
On Mon, 25 May 2026 15:38:54 -0300
Luiz Paulo Damaceno via time-nuts time-nuts@lists.febo.com wrote:
I've started to see if is possible to fix an H-maser that we have here in
our timelab. Today we're contributing to UTC With some ensemble that is
composed by: 1 commercial cs clock + BVA phase locked in long term (<100s)
to the cs + synthesizer + HROG-5. And these are results for last circular
t: Cirt 460 LRTE PPP link
https://webtai.bipm.org/ftp/pub/tai/timelinks/lkc/2604/lrteptb/lnk/lrteptb.333a_.gif
I
wish to know if this result is good for an ensemble like we have here for
starting this discussion...
Given you have only a single Cs beam standard, this looks ok and
in the order of magnitude I would expect the stability of the Cs beam.
The BVA only contributes to short term stability and thus does not
make a difference in the Circular T plots.
And what you all suggest for improvement.
This depends a lot on what you want to achieve.
Is it short term performance you are after? Is it long term stability?
Is it reliability? Is it lower uncertainty to UTC?
Quite honestly, given that you are an NMI and we are talking about
a national timescale, I would recommend that you first try to enhance
your resilience against device failures. Having only a single Cs beam
standard, which need a replacement tube every 5 to 20 years (depending
on type and chance), you run the risk of losing your timescale when your
Cs beam standard runs out of Cs.
Today i have:
Real time tropospheric delay calculation with hopfield's model to replace
"MDTR" (with also mapping function to correct for each satellite). Every 15
minutes, also, I fed the Rx5Tr receiver with temperature, humidity and
pressure for its tropospheric model update. The sensor is located near the
antenna.
At the current state, I am not sure whether GNSS time transfer is
really the limiting factor for you. At least not short term.
IIRC the current uncertainty for GNSS time transfer is around 2ns,
even with iPPP, because of delay stability in the device itself.
Of course, if you can make your antenna's filter more stable, you
would then be able to get better absolute calibration.
And an algorithm that simply sees the frequency deviation every 2 hours and
computes a frequency correction for our HROG-5.
Given your MDEV data, I would set the correction time-span to 1-2days.
At least, that's where your MDEV flattens out.
We also have a hydrogen maser that is stopped because the 10 MHz oscillator
is not locking. I want to do some maintenance there to see if it can be
brought back to life.
The maser is glowing purple, but it shows "no vacuum" and "synthesizer is
faulty". I can see an A.M. oscillation of about -140 to -120 dBm in the
cavity output, but nothing more.
So, if possible, I may want to be guided in some steps to see if is it
possible to recover it.
We have had some time-nuts members repair their H-Masers on their own.
But given that you are an NMI, I would contact the manufacturer. They
are most likely willing to help.
I hope this helps
Attila Kinali
--
The driving force behind research is the question: "Why?"
There are things we don't understand and things we always
wonder about. And that's why we do research.
-- Kobayashi Makoto
A wonderful good morning!
On Mon, 25 May 2026 15:38:54 -0300
Luiz Paulo Damaceno via time-nuts <time-nuts@lists.febo.com> wrote:
> I've started to see if is possible to fix an H-maser that we have here in
> our timelab. Today we're contributing to UTC With some ensemble that is
> composed by: 1 commercial cs clock + BVA phase locked in long term (<100s)
> to the cs + synthesizer + HROG-5. And these are results for last circular
> t: Cirt 460 LRTE PPP link
> <https://webtai.bipm.org/ftp/pub/tai/timelinks/lkc/2604/lrteptb/lnk/lrteptb.333a_.gif>
> I
> wish to know if this result is good for an ensemble like we have here for
> starting this discussion...
Given you have only a single Cs beam standard, this looks ok and
in the order of magnitude I would expect the stability of the Cs beam.
The BVA only contributes to short term stability and thus does not
make a difference in the Circular T plots.
> And what you all suggest for improvement.
This depends a lot on what you want to achieve.
Is it short term performance you are after? Is it long term stability?
Is it reliability? Is it lower uncertainty to UTC?
Quite honestly, given that you are an NMI and we are talking about
a national timescale, I would recommend that you first try to enhance
your resilience against device failures. Having only a single Cs beam
standard, which need a replacement tube every 5 to 20 years (depending
on type and chance), you run the risk of losing your timescale when your
Cs beam standard runs out of Cs.
> Today i have:
>
> Real time tropospheric delay calculation with hopfield's model to replace
> "MDTR" (with also mapping function to correct for each satellite). Every 15
> minutes, also, I fed the Rx5Tr receiver with temperature, humidity and
> pressure for its tropospheric model update. The sensor is located near the
> antenna.
At the current state, I am not sure whether GNSS time transfer is
really the limiting factor for you. At least not short term.
IIRC the current uncertainty for GNSS time transfer is around 2ns,
even with iPPP, because of delay stability in the device itself.
Of course, if you can make your antenna's filter more stable, you
would then be able to get better absolute calibration.
> And an algorithm that simply sees the frequency deviation every 2 hours and
> computes a frequency correction for our HROG-5.
Given your MDEV data, I would set the correction time-span to 1-2days.
At least, that's where your MDEV flattens out.
> We also have a hydrogen maser that is stopped because the 10 MHz oscillator
> is not locking. I want to do some maintenance there to see if it can be
> brought back to life.
>
> The maser is glowing purple, but it shows "no vacuum" and "synthesizer is
> faulty". I can see an A.M. oscillation of about -140 to -120 dBm in the
> cavity output, but nothing more.
>
> So, if possible, I may want to be guided in some steps to see if is it
> possible to recover it.
We have had some time-nuts members repair their H-Masers on their own.
But given that you are an NMI, I would contact the manufacturer. They
are most likely willing to help.
I hope this helps
Attila Kinali
--
The driving force behind research is the question: "Why?"
There are things we don't understand and things we always
wonder about. And that's why we do research.
-- Kobayashi Makoto
LP
Luiz Paulo Damaceno
Tue, May 26, 2026 1:15 PM
John,
-> The Cs+GNSS ensemble looks like it's performing very well. That's the
Septentrio PolaRx5TR receiver, correct?
That's correct.
-> It sounds like you may have a power supply problem with the maser, or
perhaps a controller issue, since you obviously have vacuum if the
dissociator is lit. How far off frequency is it? More than a few Hz
suggests that the oscillator oven has failed, which could explain everything
except the vacuum error.
Is not that too far, I can manually adjust the frequency and stability
seems quite good (unlocked PolaRx5Tr shows something around 4~3E-12 for 1s
ADEV).
Did you think checking power supply voltages (not trusting the display in
the instrument front).
Thanks
Luiz Paulo
Em ter., 26 de mai. de 2026 às 02:48, john@miles.io escreveu:
The Cs+GNSS ensemble looks like it's performing very well. That's the
Septentrio PolaRx5TR receiver, correct?
It sounds like you may have a power supply problem with the maser, or
perhaps a controller issue, since you obviously have vacuum if the
dissociator is lit. How far off frequency is it? More than a few Hz
suggests that the oscillator oven has failed, which could explain
everything
except the vacuum error.
-- john
-----Original Message-----
From: Luiz Paulo Damaceno via time-nuts time-nuts@lists.febo.com
Sent: Monday, May 25, 2026 11:39 AM
To: Discussion of precise time and frequency measurement
time-nuts@lists.febo.com
Cc: Luiz Paulo Damaceno luizpauloeletrico42@gmail.com
Subject: [time-nuts] UTC(LRTE) H-Maser maintenace
Hi all,
I've started to see if is possible to fix an H-maser that we have here in
our timelab. Today we're contributing to UTC With some ensemble that is
composed by: 1 commercial cs clock + BVA phase locked in long term (<100s)
to the cs + synthesizer + HROG-5. And these are results for last circular
t: Cirt 460 LRTE PPP link
<
https://webtai.bipm.org/ftp/pub/tai/timelinks/lkc/2604/lrteptb/lnk/lrteptb.
333a_.gif
https://webtai.bipm.org/ftp/pub/tai/timelinks/lkc/2604/lrteptb/lnk/lrteptb.333a_.gif
I
wish to know if this result is good for an ensemble like we have here for
starting this discussion... And what you all suggest for improvement. Today
i have:
Real time tropospheric delay calculation with hopfield's model to replace
"MDTR" (with also mapping function to correct for each satellite). Every 15
minutes, also, I fed the Rx5Tr receiver with temperature, humidity and
pressure for its tropospheric model update. The sensor is located near the
antenna.
And an algorithm that simply sees the frequency deviation every 2 hours and
computes a frequency correction for our HROG-5.
We also have a hydrogen maser that is stopped because the 10 MHz oscillator
is not locking. I want to do some maintenance there to see if it can be
brought back to life.
The maser is glowing purple, but it shows "no vacuum" and "synthesizer is
faulty". I can see an A.M. oscillation of about -140 to -120 dBm in the
cavity output, but nothing more.
So, if possible, I may want to be guided in some steps to see if is it
possible to recover it.
Best regards,
Luiz
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe send an
email to time-nuts-leave@lists.febo.com
John,
-> The Cs+GNSS ensemble looks like it's performing very well. That's the
Septentrio PolaRx5TR receiver, correct?
That's correct.
-> It sounds like you may have a power supply problem with the maser, or
perhaps a controller issue, since you obviously have vacuum if the
dissociator is lit. How far off frequency is it? More than a few Hz
suggests that the oscillator oven has failed, which could explain everything
except the vacuum error.
Is not that too far, I can manually adjust the frequency and stability
seems quite good (unlocked PolaRx5Tr shows something around 4~3E-12 for 1s
ADEV).
Did you think checking power supply voltages (not trusting the display in
the instrument front).
Thanks
Luiz Paulo
Em ter., 26 de mai. de 2026 às 02:48, <john@miles.io> escreveu:
> The Cs+GNSS ensemble looks like it's performing very well. That's the
> Septentrio PolaRx5TR receiver, correct?
>
> It sounds like you may have a power supply problem with the maser, or
> perhaps a controller issue, since you obviously have vacuum if the
> dissociator is lit. How far off frequency is it? More than a few Hz
> suggests that the oscillator oven has failed, which could explain
> everything
> except the vacuum error.
>
> -- john
>
> -----Original Message-----
> From: Luiz Paulo Damaceno via time-nuts <time-nuts@lists.febo.com>
> Sent: Monday, May 25, 2026 11:39 AM
> To: Discussion of precise time and frequency measurement
> <time-nuts@lists.febo.com>
> Cc: Luiz Paulo Damaceno <luizpauloeletrico42@gmail.com>
> Subject: [time-nuts] UTC(LRTE) H-Maser maintenace
>
> Hi all,
>
> I've started to see if is possible to fix an H-maser that we have here in
> our timelab. Today we're contributing to UTC With some ensemble that is
> composed by: 1 commercial cs clock + BVA phase locked in long term (<100s)
> to the cs + synthesizer + HROG-5. And these are results for last circular
> t: Cirt 460 LRTE PPP link
> <
> https://webtai.bipm.org/ftp/pub/tai/timelinks/lkc/2604/lrteptb/lnk/lrteptb.
> 333a_.gif
> <https://webtai.bipm.org/ftp/pub/tai/timelinks/lkc/2604/lrteptb/lnk/lrteptb.333a_.gif>
> >
> I
> wish to know if this result is good for an ensemble like we have here for
> starting this discussion... And what you all suggest for improvement. Today
> i have:
>
> Real time tropospheric delay calculation with hopfield's model to replace
> "MDTR" (with also mapping function to correct for each satellite). Every 15
> minutes, also, I fed the Rx5Tr receiver with temperature, humidity and
> pressure for its tropospheric model update. The sensor is located near the
> antenna.
>
> And an algorithm that simply sees the frequency deviation every 2 hours and
> computes a frequency correction for our HROG-5.
>
> We also have a hydrogen maser that is stopped because the 10 MHz oscillator
> is not locking. I want to do some maintenance there to see if it can be
> brought back to life.
>
> The maser is glowing purple, but it shows "no vacuum" and "synthesizer is
> faulty". I can see an A.M. oscillation of about -140 to -120 dBm in the
> cavity output, but nothing more.
>
> So, if possible, I may want to be guided in some steps to see if is it
> possible to recover it.
>
> Best regards,
>
> Luiz
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe send an
> email to time-nuts-leave@lists.febo.com
>
>
LP
Luiz Paulo Damaceno
Tue, May 26, 2026 1:21 PM
Thank you, Attila!
-> I want to achieve long term stability and also short term performance. I
think both are important to us. I can figure out that things better than we
have now may need more clocks or a change of base clock.
-> Quite honestly, given that you are an NMI and we are talking about
a national timescale, I would recommend that you first try to enhance
your resilience against device failures. Having only a single Cs beam
standard, which need a replacement tube every 5 to 20 years (depending
on type and chance), you run the risk of losing your timescale when your
Cs beam standard runs out of Cs.
- I'm not an NMI, I'm in a university that runs atomic physics
experiments. Our UTC time scale is one way for us to have trackability to
Cs frequency realization.
- Actually we have a second Cs beam that is not running due to this: have
a second one in case of failure, at least.
Thank you for your answers,
Luiz Paulo
Em ter., 26 de mai. de 2026 às 06:56, Attila Kinali via time-nuts <
time-nuts@lists.febo.com> escreveu:
A wonderful good morning!
On Mon, 25 May 2026 15:38:54 -0300
Luiz Paulo Damaceno via time-nuts time-nuts@lists.febo.com wrote:
I've started to see if is possible to fix an H-maser that we have here in
our timelab. Today we're contributing to UTC With some ensemble that is
composed by: 1 commercial cs clock + BVA phase locked in long term
to the cs + synthesizer + HROG-5. And these are results for last circular
t: Cirt 460 LRTE PPP link
<
I
wish to know if this result is good for an ensemble like we have here
starting this discussion...
Given you have only a single Cs beam standard, this looks ok and
in the order of magnitude I would expect the stability of the Cs beam.
The BVA only contributes to short term stability and thus does not
make a difference in the Circular T plots.
And what you all suggest for improvement.
This depends a lot on what you want to achieve.
Is it short term performance you are after? Is it long term stability?
Is it reliability? Is it lower uncertainty to UTC?
Quite honestly, given that you are an NMI and we are talking about
a national timescale, I would recommend that you first try to enhance
your resilience against device failures. Having only a single Cs beam
standard, which need a replacement tube every 5 to 20 years (depending
on type and chance), you run the risk of losing your timescale when your
Cs beam standard runs out of Cs.
Today i have:
Real time tropospheric delay calculation with hopfield's model to replace
"MDTR" (with also mapping function to correct for each satellite). Every
minutes, also, I fed the Rx5Tr receiver with temperature, humidity and
pressure for its tropospheric model update. The sensor is located near
At the current state, I am not sure whether GNSS time transfer is
really the limiting factor for you. At least not short term.
IIRC the current uncertainty for GNSS time transfer is around 2ns,
even with iPPP, because of delay stability in the device itself.
Of course, if you can make your antenna's filter more stable, you
would then be able to get better absolute calibration.
And an algorithm that simply sees the frequency deviation every 2 hours
computes a frequency correction for our HROG-5.
Given your MDEV data, I would set the correction time-span to 1-2days.
At least, that's where your MDEV flattens out.
We also have a hydrogen maser that is stopped because the 10 MHz
is not locking. I want to do some maintenance there to see if it can be
brought back to life.
The maser is glowing purple, but it shows "no vacuum" and "synthesizer is
faulty". I can see an A.M. oscillation of about -140 to -120 dBm in the
cavity output, but nothing more.
So, if possible, I may want to be guided in some steps to see if is it
possible to recover it.
We have had some time-nuts members repair their H-Masers on their own.
But given that you are an NMI, I would contact the manufacturer. They
are most likely willing to help.
I hope this helps
Attila Kinali
--
The driving force behind research is the question: "Why?"
There are things we don't understand and things we always
wonder about. And that's why we do research.
-- Kobayashi Makoto
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Thank you, Attila!
-> I want to achieve long term stability and also short term performance. I
think both are important to us. I can figure out that things better than we
have now may need more clocks or a change of base clock.
-> Quite honestly, given that you are an NMI and we are talking about
a national timescale, I would recommend that you first try to enhance
your resilience against device failures. Having only a single Cs beam
standard, which need a replacement tube every 5 to 20 years (depending
on type and chance), you run the risk of losing your timescale when your
Cs beam standard runs out of Cs.
- I'm not an NMI, I'm in a university that runs atomic physics
experiments. Our UTC time scale is one way for us to have trackability to
Cs frequency realization.
- Actually we have a second Cs beam that is not running due to this: have
a second one in case of failure, at least.
Thank you for your answers,
Luiz Paulo
Em ter., 26 de mai. de 2026 às 06:56, Attila Kinali via time-nuts <
time-nuts@lists.febo.com> escreveu:
> A wonderful good morning!
>
> On Mon, 25 May 2026 15:38:54 -0300
> Luiz Paulo Damaceno via time-nuts <time-nuts@lists.febo.com> wrote:
>
> > I've started to see if is possible to fix an H-maser that we have here in
> > our timelab. Today we're contributing to UTC With some ensemble that is
> > composed by: 1 commercial cs clock + BVA phase locked in long term
> (<100s)
> > to the cs + synthesizer + HROG-5. And these are results for last circular
> > t: Cirt 460 LRTE PPP link
> > <
> https://webtai.bipm.org/ftp/pub/tai/timelinks/lkc/2604/lrteptb/lnk/lrteptb.333a_.gif
> >
> > I
> > wish to know if this result is good for an ensemble like we have here
> for
> > starting this discussion...
>
> Given you have only a single Cs beam standard, this looks ok and
> in the order of magnitude I would expect the stability of the Cs beam.
>
> The BVA only contributes to short term stability and thus does not
> make a difference in the Circular T plots.
>
> > And what you all suggest for improvement.
>
> This depends a lot on what you want to achieve.
> Is it short term performance you are after? Is it long term stability?
> Is it reliability? Is it lower uncertainty to UTC?
>
> Quite honestly, given that you are an NMI and we are talking about
> a national timescale, I would recommend that you first try to enhance
> your resilience against device failures. Having only a single Cs beam
> standard, which need a replacement tube every 5 to 20 years (depending
> on type and chance), you run the risk of losing your timescale when your
> Cs beam standard runs out of Cs.
>
> > Today i have:
> >
> > Real time tropospheric delay calculation with hopfield's model to replace
> > "MDTR" (with also mapping function to correct for each satellite). Every
> 15
> > minutes, also, I fed the Rx5Tr receiver with temperature, humidity and
> > pressure for its tropospheric model update. The sensor is located near
> the
> > antenna.
>
> At the current state, I am not sure whether GNSS time transfer is
> really the limiting factor for you. At least not short term.
> IIRC the current uncertainty for GNSS time transfer is around 2ns,
> even with iPPP, because of delay stability in the device itself.
> Of course, if you can make your antenna's filter more stable, you
> would then be able to get better absolute calibration.
>
>
> > And an algorithm that simply sees the frequency deviation every 2 hours
> and
> > computes a frequency correction for our HROG-5.
>
> Given your MDEV data, I would set the correction time-span to 1-2days.
> At least, that's where your MDEV flattens out.
>
> > We also have a hydrogen maser that is stopped because the 10 MHz
> oscillator
> > is not locking. I want to do some maintenance there to see if it can be
> > brought back to life.
> >
> > The maser is glowing purple, but it shows "no vacuum" and "synthesizer is
> > faulty". I can see an A.M. oscillation of about -140 to -120 dBm in the
> > cavity output, but nothing more.
> >
> > So, if possible, I may want to be guided in some steps to see if is it
> > possible to recover it.
>
> We have had some time-nuts members repair their H-Masers on their own.
> But given that you are an NMI, I would contact the manufacturer. They
> are most likely willing to help.
>
> I hope this helps
>
> Attila Kinali
>
> --
> The driving force behind research is the question: "Why?"
> There are things we don't understand and things we always
> wonder about. And that's why we do research.
> -- Kobayashi Makoto
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
>
AK
Attila Kinali
Tue, May 26, 2026 2:27 PM
-> I want to achieve long term stability and also short term performance.
I think both are important to us. I can figure out that things better than we
have now may need more clocks or a change of base clock.
Ok. One "easy" first thing you can do is to more loosely couple
your BVA to your Cs beam standard. If your BVA is well aged,
it is likely to perform better than the Cs beam standard up
to 1000s. If you change the loop-bandwidth of your PLL such
that you get somewhere between 0.01Hz and 0.001Hz, then you
could make better use of your BVA. Of course, that's not
as easy as it sounds. If you have some phase measurement
equipment (e.g. a Timepod, but also a red pitaya would probably
do), you could use the BVA as input to the HROG, then compare
the phase of the HROG output and the Cs beam standard. It would
be then pretty easy to do the PLL loop on some PC and use the
HROG for the control/feedback part.
-> Quite honestly, given that you are an NMI and we are talking about
a national timescale, I would recommend that you first try to enhance
your resilience against device failures. Having only a single Cs beam
standard, which need a replacement tube every 5 to 20 years (depending
on type and chance), you run the risk of losing your timescale when your
Cs beam standard runs out of Cs.
- I'm not an NMI, I'm in a university that runs atomic physics
experiments.
Oh.. sorry. My apologies! I assumed that you were an NMI as it is
rare that non-NMIs contribute to TAI/UTC.
Our UTC time scale is one way for us to have trackability to
Cs frequency realization.
- Actually we have a second Cs beam that is not running due to this: have
a second one in case of failure, at least.
Ok, so you don't have any legal requirement to keep a continuous
timescale. I guess your primary need is to provide a local realization
of the SI second that is as accurate as possible with as little noise
and uncertainty as possible?
Then I would focus on getting the hydrogen maser working again
and use it as input to the HROG. That will provide you the best
short to medium term performance.
Then use your phase meter to compare your HROG output and the
Cs beam output. This way you can correct for the drift of the
maser. Use the GNSS system to correct the drift of the Cs beam
standard with a tau in the range of a few weeks to a month.
This is basically what METAS does (minus the Cs fountain that
they periodically run) [1].
BTW: If your research involves optical atomic clocks and their
absolute frequency measurement, you might want to consider getting
a primary frequency standard. In the course of the presentation on the
2025 BIPM recommended frequency list at EFTF last month, Helen Margolis
from NPL did complain about the correlations of absolute frequency
measurements done through UTC, which are very hard to properly handle.
Attila Kinali
[1] "METAS new time scale generation system - A progress report"
by Bernier, Dudle and Schlunegger, 2006
http://time.kinali.ch/ptti/2006papers/paper3.pdf
The driving force behind research is the question: "Why?"
There are things we don't understand and things we always
wonder about. And that's why we do research.
-- Kobayashi Makoto
Ola Luiz Paulo!
On Tue, 26 May 2026 10:21:03 -0300
Luiz Paulo Damaceno <luizpauloeletrico42@gmail.com> wrote:
> -> I want to achieve long term stability and also short term performance.
> I think both are important to us. I can figure out that things better than we
> have now may need more clocks or a change of base clock.
Ok. One "easy" first thing you can do is to more loosely couple
your BVA to your Cs beam standard. If your BVA is well aged,
it is likely to perform better than the Cs beam standard up
to 1000s. If you change the loop-bandwidth of your PLL such
that you get somewhere between 0.01Hz and 0.001Hz, then you
could make better use of your BVA. Of course, that's not
as easy as it sounds. If you have some phase measurement
equipment (e.g. a Timepod, but also a red pitaya would probably
do), you could use the BVA as input to the HROG, then compare
the phase of the HROG output and the Cs beam standard. It would
be then pretty easy to do the PLL loop on some PC and use the
HROG for the control/feedback part.
> -> Quite honestly, given that you are an NMI and we are talking about
> a national timescale, I would recommend that you first try to enhance
> your resilience against device failures. Having only a single Cs beam
> standard, which need a replacement tube every 5 to 20 years (depending
> on type and chance), you run the risk of losing your timescale when your
> Cs beam standard runs out of Cs.
> - I'm not an NMI, I'm in a university that runs atomic physics
> experiments.
Oh.. sorry. My apologies! I assumed that you were an NMI as it is
rare that non-NMIs contribute to TAI/UTC.
> Our UTC time scale is one way for us to have trackability to
> Cs frequency realization.
> - Actually we have a second Cs beam that is not running due to this: have
> a second one in case of failure, at least.
Ok, so you don't have any legal requirement to keep a continuous
timescale. I guess your primary need is to provide a local realization
of the SI second that is as accurate as possible with as little noise
and uncertainty as possible?
Then I would focus on getting the hydrogen maser working again
and use it as input to the HROG. That will provide you the best
short to medium term performance.
Then use your phase meter to compare your HROG output and the
Cs beam output. This way you can correct for the drift of the
maser. Use the GNSS system to correct the drift of the Cs beam
standard with a tau in the range of a few weeks to a month.
This is basically what METAS does (minus the Cs fountain that
they periodically run) [1].
BTW: If your research involves optical atomic clocks and their
absolute frequency measurement, you might want to consider getting
a primary frequency standard. In the course of the presentation on the
2025 BIPM recommended frequency list at EFTF last month, Helen Margolis
from NPL did complain about the correlations of absolute frequency
measurements done through UTC, which are very hard to properly handle.
Attila Kinali
[1] "METAS new time scale generation system - A progress report"
by Bernier, Dudle and Schlunegger, 2006
http://time.kinali.ch/ptti/2006papers/paper3.pdf
--
The driving force behind research is the question: "Why?"
There are things we don't understand and things we always
wonder about. And that's why we do research.
-- Kobayashi Makoto
J
john@miles.io
Tue, May 26, 2026 4:16 PM
Hard to say without knowing anything about the unit in question, but it’s not a bad idea to look at the supply rails with an actual DMM or scope. Depending on resolution, sampling rate and smoothing, ripple or noise/instability may not get flagged by the self-monitoring hardware.
I think I’d focus on the vacuum failure message after that. Is it just complaining that the ion pump current is slightly high, or is it actually reading a separate vacuum gauge and not liking what it sees? A false report might be a clue to the same problem that keeps the control loop from locking. At the same time, maybe the vacuum integrity is good enough to allow the H2 dissociator to light up, but not good enough to support maser action.
If it’s an active maser and you are seeing any signal at all from the cavity, that sounds like a good sign. If it’s a passive maser, there will always be some 1420 MHz output as long as the input port is driven, so that wouldn’t necessarily tell you much.
-- john
From: Luiz Paulo Damaceno luizpauloeletrico42@gmail.com
Is not that too far, I can manually adjust the frequency and stability seems quite good (unlocked PolaRx5Tr shows something around 4~3E-12 for 1s ADEV).
Did you think checking power supply voltages (not trusting the display in the instrument front).
Hard to say without knowing anything about the unit in question, but it’s not a bad idea to look at the supply rails with an actual DMM or scope. Depending on resolution, sampling rate and smoothing, ripple or noise/instability may not get flagged by the self-monitoring hardware.
I think I’d focus on the vacuum failure message after that. Is it just complaining that the ion pump current is slightly high, or is it actually reading a separate vacuum gauge and not liking what it sees? A false report might be a clue to the same problem that keeps the control loop from locking. At the same time, maybe the vacuum integrity is good enough to allow the H2 dissociator to light up, but not good enough to support maser action.
If it’s an active maser and you are seeing any signal at all from the cavity, that sounds like a good sign. If it’s a passive maser, there will always be some 1420 MHz output as long as the input port is driven, so that wouldn’t necessarily tell you much.
-- john
From: Luiz Paulo Damaceno <luizpauloeletrico42@gmail.com>
Is not that too far, I can manually adjust the frequency and stability seems quite good (unlocked PolaRx5Tr shows something around 4~3E-12 for 1s ADEV).
Did you think checking power supply voltages (not trusting the display in the instrument front).
LP
Luiz Paulo Damaceno
Thu, May 28, 2026 1:37 PM
Ok. One "easy" first thing you can do is to more loosely couple
your BVA to your Cs beam standard. If your BVA is well aged,
it is likely to perform better than the Cs beam standard up
to 1000s. If you change the loop-bandwidth of your PLL such
that you get somewhere between 0.01Hz and 0.001Hz, then you
could make better use of your BVA. Of course, that's not
as easy as it sounds. If you have some phase measurement
equipment (e.g. a Timepod, but also a red pitaya would probably
do), you could use the BVA as input to the HROG, then compare
the phase of the HROG output and the Cs beam standard. It would
be then pretty easy to do the PLL loop on some PC and use the
HROG for the control/feedback part.
Thank you for the considerations. I've tunned the PLL for 0.00289Hz, today
i'm using a 1,1M ohm resistance and 50 uF capacitor. I can also reduce the
error voltage by changing the voltage in a potentiometer, just like a
"volume" control in audio systems. I expect to reduce a bit 0,25 times day
to day) to archive 1000s loop. I can only evaluate this using GNSS PPP - by
RTKLib in real time or NRCAN for example and / or CGGTTS files.
I have also available for these measurements a frequency comb from menlo
systems. Can I use repetition rate adev to evaluate the BVA? (locking it in
a Cs beam and then in BVA + Cs beam)?
Ok, so you don't have any legal requirement to keep a continuous
timescale. I guess your primary need is to provide a local realization
of the SI second that is as accurate as possible with as little noise
and uncertainty as possible?
Yes, no legal requirement, here in Brazil, only ONJR has the legal
requirement. Yes, my primary need is to provide a local realization of SI
second with little noise and uncertainty.
We are working in cold atom clocks here also. A Cs fountain and another
kind of cold atom Cs clock. But now, we're in the construction phase. No
measurements being made.
About getting the maser working again, do you have any experience with that
maser? Is the KVARZ CH1-75A.
Abraços,
Luiz Paulo
Em ter., 26 de mai. de 2026 às 11:27, Attila Kinali attila@kinali.ch
escreveu:
-> I want to achieve long term stability and also short term
I think both are important to us. I can figure out that things better
have now may need more clocks or a change of base clock.
Ok. One "easy" first thing you can do is to more loosely couple
your BVA to your Cs beam standard. If your BVA is well aged,
it is likely to perform better than the Cs beam standard up
to 1000s. If you change the loop-bandwidth of your PLL such
that you get somewhere between 0.01Hz and 0.001Hz, then you
could make better use of your BVA. Of course, that's not
as easy as it sounds. If you have some phase measurement
equipment (e.g. a Timepod, but also a red pitaya would probably
do), you could use the BVA as input to the HROG, then compare
the phase of the HROG output and the Cs beam standard. It would
be then pretty easy to do the PLL loop on some PC and use the
HROG for the control/feedback part.
-> Quite honestly, given that you are an NMI and we are talking about
a national timescale, I would recommend that you first try to enhance
your resilience against device failures. Having only a single Cs beam
standard, which need a replacement tube every 5 to 20 years (depending
on type and chance), you run the risk of losing your timescale when your
Cs beam standard runs out of Cs.
- I'm not an NMI, I'm in a university that runs atomic physics
experiments.
Oh.. sorry. My apologies! I assumed that you were an NMI as it is
rare that non-NMIs contribute to TAI/UTC.
Our UTC time scale is one way for us to have trackability to
Cs frequency realization.
- Actually we have a second Cs beam that is not running due to this:
a second one in case of failure, at least.
Ok, so you don't have any legal requirement to keep a continuous
timescale. I guess your primary need is to provide a local realization
of the SI second that is as accurate as possible with as little noise
and uncertainty as possible?
Then I would focus on getting the hydrogen maser working again
and use it as input to the HROG. That will provide you the best
short to medium term performance.
Then use your phase meter to compare your HROG output and the
Cs beam output. This way you can correct for the drift of the
maser. Use the GNSS system to correct the drift of the Cs beam
standard with a tau in the range of a few weeks to a month.
This is basically what METAS does (minus the Cs fountain that
they periodically run) [1].
BTW: If your research involves optical atomic clocks and their
absolute frequency measurement, you might want to consider getting
a primary frequency standard. In the course of the presentation on the
2025 BIPM recommended frequency list at EFTF last month, Helen Margolis
from NPL did complain about the correlations of absolute frequency
measurements done through UTC, which are very hard to properly handle.
Attila Kinali
[1] "METAS new time scale generation system - A progress report"
by Bernier, Dudle and Schlunegger, 2006
http://time.kinali.ch/ptti/2006papers/paper3.pdf
The driving force behind research is the question: "Why?"
There are things we don't understand and things we always
wonder about. And that's why we do research.
-- Kobayashi Makoto
Ola, Attila!
>Ok. One "easy" first thing you can do is to more loosely couple
>your BVA to your Cs beam standard. If your BVA is well aged,
>it is likely to perform better than the Cs beam standard up
>to 1000s. If you change the loop-bandwidth of your PLL such
>that you get somewhere between 0.01Hz and 0.001Hz, then you
>could make better use of your BVA. Of course, that's not
>as easy as it sounds. If you have some phase measurement
>equipment (e.g. a Timepod, but also a red pitaya would probably
>do), you could use the BVA as input to the HROG, then compare
>the phase of the HROG output and the Cs beam standard. It would
>be then pretty easy to do the PLL loop on some PC and use the
>HROG for the control/feedback part.
Thank you for the considerations. I've tunned the PLL for 0.00289Hz, today
i'm using a 1,1M ohm resistance and 50 uF capacitor. I can also reduce the
error voltage by changing the voltage in a potentiometer, just like a
"volume" control in audio systems. I expect to reduce a bit 0,25 times day
to day) to archive 1000s loop. I can only evaluate this using GNSS PPP - by
RTKLib in real time or NRCAN for example and / or CGGTTS files.
I have also available for these measurements a frequency comb from menlo
systems. Can I use repetition rate adev to evaluate the BVA? (locking it in
a Cs beam and then in BVA + Cs beam)?
>Ok, so you don't have any legal requirement to keep a continuous
>timescale. I guess your primary need is to provide a local realization
>of the SI second that is as accurate as possible with as little noise
>and uncertainty as possible?
Yes, no legal requirement, here in Brazil, only ONJR has the legal
requirement. Yes, my primary need is to provide a local realization of SI
second with little noise and uncertainty.
We are working in cold atom clocks here also. A Cs fountain and another
kind of cold atom Cs clock. But now, we're in the construction phase. No
measurements being made.
About getting the maser working again, do you have any experience with that
maser? Is the KVARZ CH1-75A.
Abraços,
Luiz Paulo
Em ter., 26 de mai. de 2026 às 11:27, Attila Kinali <attila@kinali.ch>
escreveu:
> Ola Luiz Paulo!
>
> On Tue, 26 May 2026 10:21:03 -0300
> Luiz Paulo Damaceno <luizpauloeletrico42@gmail.com> wrote:
>
> > -> I want to achieve long term stability and also short term
> performance.
> > I think both are important to us. I can figure out that things better
> than we
> > have now may need more clocks or a change of base clock.
>
> Ok. One "easy" first thing you can do is to more loosely couple
> your BVA to your Cs beam standard. If your BVA is well aged,
> it is likely to perform better than the Cs beam standard up
> to 1000s. If you change the loop-bandwidth of your PLL such
> that you get somewhere between 0.01Hz and 0.001Hz, then you
> could make better use of your BVA. Of course, that's not
> as easy as it sounds. If you have some phase measurement
> equipment (e.g. a Timepod, but also a red pitaya would probably
> do), you could use the BVA as input to the HROG, then compare
> the phase of the HROG output and the Cs beam standard. It would
> be then pretty easy to do the PLL loop on some PC and use the
> HROG for the control/feedback part.
>
> > -> Quite honestly, given that you are an NMI and we are talking about
> > a national timescale, I would recommend that you first try to enhance
> > your resilience against device failures. Having only a single Cs beam
> > standard, which need a replacement tube every 5 to 20 years (depending
> > on type and chance), you run the risk of losing your timescale when your
> > Cs beam standard runs out of Cs.
> > - I'm not an NMI, I'm in a university that runs atomic physics
> > experiments.
>
> Oh.. sorry. My apologies! I assumed that you were an NMI as it is
> rare that non-NMIs contribute to TAI/UTC.
>
> > Our UTC time scale is one way for us to have trackability to
> > Cs frequency realization.
> > - Actually we have a second Cs beam that is not running due to this:
> have
> > a second one in case of failure, at least.
>
> Ok, so you don't have any legal requirement to keep a continuous
> timescale. I guess your primary need is to provide a local realization
> of the SI second that is as accurate as possible with as little noise
> and uncertainty as possible?
>
> Then I would focus on getting the hydrogen maser working again
> and use it as input to the HROG. That will provide you the best
> short to medium term performance.
>
> Then use your phase meter to compare your HROG output and the
> Cs beam output. This way you can correct for the drift of the
> maser. Use the GNSS system to correct the drift of the Cs beam
> standard with a tau in the range of a few weeks to a month.
>
> This is basically what METAS does (minus the Cs fountain that
> they periodically run) [1].
>
> BTW: If your research involves optical atomic clocks and their
> absolute frequency measurement, you might want to consider getting
> a primary frequency standard. In the course of the presentation on the
> 2025 BIPM recommended frequency list at EFTF last month, Helen Margolis
> from NPL did complain about the correlations of absolute frequency
> measurements done through UTC, which are very hard to properly handle.
>
>
> Attila Kinali
>
>
> [1] "METAS new time scale generation system - A progress report"
> by Bernier, Dudle and Schlunegger, 2006
> http://time.kinali.ch/ptti/2006papers/paper3.pdf
> --
> The driving force behind research is the question: "Why?"
> There are things we don't understand and things we always
> wonder about. And that's why we do research.
> -- Kobayashi Makoto
>
LP
Luiz Paulo Damaceno
Thu, May 28, 2026 1:39 PM
Thank you john!
I will verify supply rails with DMM / scope and return the results. For
now, Is it an active maser unit. CH1-75A from KVARZ.
Kind regards,
Luiz Paulo
Em ter., 26 de mai. de 2026 às 13:16, john@miles.io escreveu:
Hard to say without knowing anything about the unit in question, but it’s
not a bad idea to look at the supply rails with an actual DMM or scope.
Depending on resolution, sampling rate and smoothing, ripple or
noise/instability may not get flagged by the self-monitoring hardware.
I think I’d focus on the vacuum failure message after that. Is it just
complaining that the ion pump current is slightly high, or is it actually
reading a separate vacuum gauge and not liking what it sees? A false
report might be a clue to the same problem that keeps the control loop from
locking. At the same time, maybe the vacuum integrity is good enough to
allow the H2 dissociator to light up, but not good enough to support maser
action.
If it’s an active maser and you are seeing any signal at all from the
cavity, that sounds like a good sign. If it’s a passive maser, there will
always be some 1420 MHz output as long as the input port is driven, so that
wouldn’t necessarily tell you much.
-- john
From: Luiz Paulo Damaceno luizpauloeletrico42@gmail.com
Is not that too far, I can manually adjust the frequency and stability
seems quite good (unlocked PolaRx5Tr shows something around 4~3E-12 for 1s
ADEV).
Did you think checking power supply voltages (not trusting the display in
the instrument front).
Thank you john!
I will verify supply rails with DMM / scope and return the results. For
now, Is it an active maser unit. CH1-75A from KVARZ.
Kind regards,
Luiz Paulo
Em ter., 26 de mai. de 2026 às 13:16, <john@miles.io> escreveu:
> Hard to say without knowing anything about the unit in question, but it’s
> not a bad idea to look at the supply rails with an actual DMM or scope.
> Depending on resolution, sampling rate and smoothing, ripple or
> noise/instability may not get flagged by the self-monitoring hardware.
>
>
>
> I think I’d focus on the vacuum failure message after that. Is it just
> complaining that the ion pump current is slightly high, or is it actually
> reading a separate vacuum gauge and not liking what it sees? A false
> report might be a clue to the same problem that keeps the control loop from
> locking. At the same time, maybe the vacuum integrity is good enough to
> allow the H2 dissociator to light up, but not good enough to support maser
> action.
>
>
>
> If it’s an active maser and you are seeing any signal at all from the
> cavity, that sounds like a good sign. If it’s a passive maser, there will
> always be some 1420 MHz output as long as the input port is driven, so that
> wouldn’t necessarily tell you much.
>
>
>
> -- john
>
>
>
>
>
> *From:* Luiz Paulo Damaceno <luizpauloeletrico42@gmail.com>
>
> Is not that too far, I can manually adjust the frequency and stability
> seems quite good (unlocked PolaRx5Tr shows something around 4~3E-12 for 1s
> ADEV).
>
> Did you think checking power supply voltages (not trusting the display in
> the instrument front).
>
AK
Attila Kinali
Thu, May 28, 2026 3:30 PM
A wonderful good afternoon,
On Thu, 28 May 2026 10:37:58 -0300
Luiz Paulo Damaceno via time-nuts time-nuts@lists.febo.com wrote:
Thank you for the considerations. I've tunned the PLL for 0.00289Hz, today
i'm using a 1,1M ohm resistance and 50 uF capacitor. I can also reduce the
error voltage by changing the voltage in a potentiometer, just like a
"volume" control in audio systems. I expect to reduce a bit 0,25 times day
to day) to archive 1000s loop.
A PLL using the EFC input of the BVA will not work well at these
long time constants. About 1Hz loop bandwidth is the lowest that's
reasonable, I would even draw the line at 10Hz. And only if you
build an active filter using an low-noise opamp. Going lower will
limit your oscillators stability by the noise on the EFC voltage.
You have a 1MΩ resistor, which already produces quite a bit of
noise voltage. And then you use a 50µF capacitor. If it's an
electrolytic capacitor, then the leakage current noise will will destroy
the stability. If it's a ceramic capacitor, then its piezoelectric
properties will couple any acoustic noise and vibration into your
EFC voltage.
Best to ground the EFC voltage and then use a phase/frequency
stepper to shift the frequency of the BVA using a digital control
loop.-
I can only evaluate this using GNSS PPP - by
RTKLib in real time or NRCAN for example and / or CGGTTS files.
I have also available for these measurements a frequency comb from menlo
systems. Can I use repetition rate adev to evaluate the BVA? (locking it in
a Cs beam and then in BVA + Cs beam)?
If the frequency comb is CEO stabilized (aka f-2f lock) and you can
stabilize its repetition rate using an optical cavity (10cm ULE is
already enough), then you can get better stability than your BVA
in the range up to 10s probably better than 100s, depending on your
setup. If the comb is free running, then the intrinsic instability
of the comb will be larger than the BVA.
We are working in cold atom clocks here also. A Cs fountain and another
kind of cold atom Cs clock. But now, we're in the construction phase. No
measurements being made.
Oh nice! Do you already have any publications on this?
I could only find [1] and [2].
BTW: If you want some known-good reference. The Canadians (IIRC the
NRC itself) sell caesium fountains for a reasonable price.
About getting the maser working again, do you have any experience with that
maser? Is the KVARZ CH1-75A.
I have never personally worked on a AHM myself. I do know their
working principle and how the electronics is supposed to look.
Unfortunately, I don't know anyone at Kvarz, only at the former
T4-Science.
Attila Kinali
[1] "Progress in Brazilian Cesium atomic fountain-BrCsF",
by Rodriguez, Bagnato, Bueno et al, 2016
https://doi.org/10.1109/FCS.2016.7546818
[2] "Progress toward Brazilian cesium fountain second generation",
by Bueno, Rodriguez, Müller, et al. 2018
http://dx.doi.org/10.1088/1742-6596/975/1/012071
--
The driving force behind research is the question: "Why?"
There are things we don't understand and things we always
wonder about. And that's why we do research.
-- Kobayashi Makoto
A wonderful good afternoon,
On Thu, 28 May 2026 10:37:58 -0300
Luiz Paulo Damaceno via time-nuts <time-nuts@lists.febo.com> wrote:
> Thank you for the considerations. I've tunned the PLL for 0.00289Hz, today
> i'm using a 1,1M ohm resistance and 50 uF capacitor. I can also reduce the
> error voltage by changing the voltage in a potentiometer, just like a
> "volume" control in audio systems. I expect to reduce a bit 0,25 times day
> to day) to archive 1000s loop.
A PLL using the EFC input of the BVA will not work well at these
long time constants. About 1Hz loop bandwidth is the lowest that's
reasonable, I would even draw the line at 10Hz. And only if you
build an active filter using an low-noise opamp. Going lower will
limit your oscillators stability by the noise on the EFC voltage.
You have a 1MΩ resistor, which already produces quite a bit of
noise voltage. And then you use a 50µF capacitor. If it's an
electrolytic capacitor, then the leakage current noise will will destroy
the stability. If it's a ceramic capacitor, then its piezoelectric
properties will couple any acoustic noise and vibration into your
EFC voltage.
Best to ground the EFC voltage and then use a phase/frequency
stepper to shift the frequency of the BVA using a digital control
loop.-
> I can only evaluate this using GNSS PPP - by
> RTKLib in real time or NRCAN for example and / or CGGTTS files.
>
> I have also available for these measurements a frequency comb from menlo
> systems. Can I use repetition rate adev to evaluate the BVA? (locking it in
> a Cs beam and then in BVA + Cs beam)?
If the frequency comb is CEO stabilized (aka f-2f lock) and you can
stabilize its repetition rate using an optical cavity (10cm ULE is
already enough), then you can get better stability than your BVA
in the range up to 10s probably better than 100s, depending on your
setup. If the comb is free running, then the intrinsic instability
of the comb will be larger than the BVA.
> We are working in cold atom clocks here also. A Cs fountain and another
> kind of cold atom Cs clock. But now, we're in the construction phase. No
> measurements being made.
Oh nice! Do you already have any publications on this?
I could only find [1] and [2].
BTW: If you want some known-good reference. The Canadians (IIRC the
NRC itself) sell caesium fountains for a reasonable price.
> About getting the maser working again, do you have any experience with that
> maser? Is the KVARZ CH1-75A.
I have never personally worked on a AHM myself. I do know their
working principle and how the electronics is supposed to look.
Unfortunately, I don't know anyone at Kvarz, only at the former
T4-Science.
Attila Kinali
[1] "Progress in Brazilian Cesium atomic fountain-BrCsF",
by Rodriguez, Bagnato, Bueno et al, 2016
https://doi.org/10.1109/FCS.2016.7546818
[2] "Progress toward Brazilian cesium fountain second generation",
by Bueno, Rodriguez, Müller, et al. 2018
http://dx.doi.org/10.1088/1742-6596/975/1/012071
--
The driving force behind research is the question: "Why?"
There are things we don't understand and things we always
wonder about. And that's why we do research.
-- Kobayashi Makoto
J
john@miles.io
Thu, May 28, 2026 4:36 PM
There is probably not much wrong with it, then, if the cavity is putting out a signal.
One thing that can cause some confusion with the Russian masers is degradation of the coaxial capacitor in the HFO. Both Tom and I have run into that before. It operates with a high RF voltage across it, and if the insulation breaks down and starts arcing intermittently, some significant EMI can occur before it gets bad enough to stop oscillating completely. The controller can potentially be affected by that.
-- john
From: Luiz Paulo Damaceno luizpauloeletrico42@gmail.com
Sent: Thursday, May 28, 2026 6:40 AM
To: john@miles.io
Cc: Discussion of precise time and frequency measurement time-nuts@lists.febo.com
Subject: Re: [time-nuts] UTC(LRTE) H-Maser maintenace
Thank you john!
I will verify supply rails with DMM / scope and return the results. For now, Is it an active maser unit. CH1-75A from KVARZ.
Kind regards,
Luiz Paulo
There is probably not much wrong with it, then, if the cavity is putting out a signal.
One thing that can cause some confusion with the Russian masers is degradation of the coaxial capacitor in the HFO. Both Tom and I have run into that before. It operates with a high RF voltage across it, and if the insulation breaks down and starts arcing intermittently, some significant EMI can occur before it gets bad enough to stop oscillating completely. The controller can potentially be affected by that.
-- john
From: Luiz Paulo Damaceno <luizpauloeletrico42@gmail.com>
Sent: Thursday, May 28, 2026 6:40 AM
To: john@miles.io
Cc: Discussion of precise time and frequency measurement <time-nuts@lists.febo.com>
Subject: Re: [time-nuts] UTC(LRTE) H-Maser maintenace
Thank you john!
I will verify supply rails with DMM / scope and return the results. For now, Is it an active maser unit. CH1-75A from KVARZ.
Kind regards,
Luiz Paulo
LP
Luiz Paulo Damaceno
Thu, May 28, 2026 6:13 PM
A PLL using the EFC input of the BVA will not work well at these
long time constants. About 1Hz loop bandwidth is the lowest that's
reasonable, I would even draw the line at 10Hz. And only if you
build an active filter using an low-noise opamp. Going lower will
limit your oscillators stability by the noise on the EFC voltage.
You have a 1MΩ resistor, which already produces quite a bit of
noise voltage. And then you use a 50µF capacitor. If it's an
electrolytic capacitor, then the leakage current noise will will destroy
the stability. If it's a ceramic capacitor, then its piezoelectric
properties will couple any acoustic noise and vibration into your
EFC voltage.
Best to ground the EFC voltage and then use a phase/frequency
stepper to shift the frequency of the BVA using a digital control
loop.-
I Did tought that even low noise components that I have here (resistors,
capacitor opamps (like OPA637 and OP37/27)) may contour these issues. Also,
the PLL circuit is temperature-stabilized and mechanical noise poorly
insulated. But you have a good point. I may use some SG384 that I also have
here to remove the frequency offset and actuate in the long term drift to
lock it with the cesium.
If the frequency comb is CEO stabilized (aka f-2f lock) and you can
stabilize its repetition rate using an optical cavity (10cm ULE is
already enough), then you can get better stability than your BVA
in the range up to 10s probably better than 100s, depending on your
setup. If the comb is free running, then the intrinsic instability
of the comb will be larger than the BVA.
We also have an optical cavity that is in progress to lock CEO. But this
will take a while I guess. The comb is repetition rate locked to BVA and
locked itself in the CEO.
Oh nice! Do you already have any publications on this?
I could only find [1] and [2].
Unfortunately that's only the both we have for now... We're ran out of
people working on this and haven't publications for a while.
Luiz Paulo
Em qui., 28 de mai. de 2026 às 12:30, Attila Kinali attila@kinali.ch
escreveu:
A wonderful good afternoon,
On Thu, 28 May 2026 10:37:58 -0300
Luiz Paulo Damaceno via time-nuts time-nuts@lists.febo.com wrote:
Thank you for the considerations. I've tunned the PLL for 0.00289Hz,
i'm using a 1,1M ohm resistance and 50 uF capacitor. I can also reduce
error voltage by changing the voltage in a potentiometer, just like a
"volume" control in audio systems. I expect to reduce a bit 0,25 times
to day) to archive 1000s loop.
A PLL using the EFC input of the BVA will not work well at these
long time constants. About 1Hz loop bandwidth is the lowest that's
reasonable, I would even draw the line at 10Hz. And only if you
build an active filter using an low-noise opamp. Going lower will
limit your oscillators stability by the noise on the EFC voltage.
You have a 1MΩ resistor, which already produces quite a bit of
noise voltage. And then you use a 50µF capacitor. If it's an
electrolytic capacitor, then the leakage current noise will will destroy
the stability. If it's a ceramic capacitor, then its piezoelectric
properties will couple any acoustic noise and vibration into your
EFC voltage.
Best to ground the EFC voltage and then use a phase/frequency
stepper to shift the frequency of the BVA using a digital control
loop.-
I can only evaluate this using GNSS PPP - by
RTKLib in real time or NRCAN for example and / or CGGTTS files.
I have also available for these measurements a frequency comb from menlo
systems. Can I use repetition rate adev to evaluate the BVA? (locking it
a Cs beam and then in BVA + Cs beam)?
If the frequency comb is CEO stabilized (aka f-2f lock) and you can
stabilize its repetition rate using an optical cavity (10cm ULE is
already enough), then you can get better stability than your BVA
in the range up to 10s probably better than 100s, depending on your
setup. If the comb is free running, then the intrinsic instability
of the comb will be larger than the BVA.
We are working in cold atom clocks here also. A Cs fountain and another
kind of cold atom Cs clock. But now, we're in the construction phase. No
measurements being made.
Oh nice! Do you already have any publications on this?
I could only find [1] and [2].
BTW: If you want some known-good reference. The Canadians (IIRC the
NRC itself) sell caesium fountains for a reasonable price.
About getting the maser working again, do you have any experience with
maser? Is the KVARZ CH1-75A.
I have never personally worked on a AHM myself. I do know their
working principle and how the electronics is supposed to look.
Unfortunately, I don't know anyone at Kvarz, only at the former
T4-Science.
Attila Kinali
[1] "Progress in Brazilian Cesium atomic fountain-BrCsF",
by Rodriguez, Bagnato, Bueno et al, 2016
https://doi.org/10.1109/FCS.2016.7546818
[2] "Progress toward Brazilian cesium fountain second generation",
by Bueno, Rodriguez, Müller, et al. 2018
http://dx.doi.org/10.1088/1742-6596/975/1/012071
--
The driving force behind research is the question: "Why?"
There are things we don't understand and things we always
wonder about. And that's why we do research.
-- Kobayashi Makoto
Attila,
>A PLL using the EFC input of the BVA will not work well at these
>long time constants. About 1Hz loop bandwidth is the lowest that's
>reasonable, I would even draw the line at 10Hz. And only if you
>build an active filter using an low-noise opamp. Going lower will
>limit your oscillators stability by the noise on the EFC voltage.
>You have a 1MΩ resistor, which already produces quite a bit of
>noise voltage. And then you use a 50µF capacitor. If it's an
>electrolytic capacitor, then the leakage current noise will will destroy
>the stability. If it's a ceramic capacitor, then its piezoelectric
>properties will couple any acoustic noise and vibration into your
>EFC voltage.
>Best to ground the EFC voltage and then use a phase/frequency
>stepper to shift the frequency of the BVA using a digital control
>loop.-
I Did tought that even low noise components that I have here (resistors,
capacitor opamps (like OPA637 and OP37/27)) may contour these issues. Also,
the PLL circuit is temperature-stabilized and mechanical noise poorly
insulated. But you have a good point. I may use some SG384 that I also have
here to remove the frequency offset and actuate in the long term drift to
lock it with the cesium.
>If the frequency comb is CEO stabilized (aka f-2f lock) and you can
>stabilize its repetition rate using an optical cavity (10cm ULE is
>already enough), then you can get better stability than your BVA
>in the range up to 10s probably better than 100s, depending on your
>setup. If the comb is free running, then the intrinsic instability
>of the comb will be larger than the BVA.
We also have an optical cavity that is in progress to lock CEO. But this
will take a while I guess. The comb is repetition rate locked to BVA and
locked itself in the CEO.
>Oh nice! Do you already have any publications on this?
>I could only find [1] and [2].
Unfortunately that's only the both we have for now... We're ran out of
people working on this and haven't publications for a while.
Luiz Paulo
Em qui., 28 de mai. de 2026 às 12:30, Attila Kinali <attila@kinali.ch>
escreveu:
> A wonderful good afternoon,
>
> On Thu, 28 May 2026 10:37:58 -0300
> Luiz Paulo Damaceno via time-nuts <time-nuts@lists.febo.com> wrote:
>
> > Thank you for the considerations. I've tunned the PLL for 0.00289Hz,
> today
> > i'm using a 1,1M ohm resistance and 50 uF capacitor. I can also reduce
> the
> > error voltage by changing the voltage in a potentiometer, just like a
> > "volume" control in audio systems. I expect to reduce a bit 0,25 times
> day
> > to day) to archive 1000s loop.
>
> A PLL using the EFC input of the BVA will not work well at these
> long time constants. About 1Hz loop bandwidth is the lowest that's
> reasonable, I would even draw the line at 10Hz. And only if you
> build an active filter using an low-noise opamp. Going lower will
> limit your oscillators stability by the noise on the EFC voltage.
> You have a 1MΩ resistor, which already produces quite a bit of
> noise voltage. And then you use a 50µF capacitor. If it's an
> electrolytic capacitor, then the leakage current noise will will destroy
> the stability. If it's a ceramic capacitor, then its piezoelectric
> properties will couple any acoustic noise and vibration into your
> EFC voltage.
>
> Best to ground the EFC voltage and then use a phase/frequency
> stepper to shift the frequency of the BVA using a digital control
> loop.-
>
> > I can only evaluate this using GNSS PPP - by
> > RTKLib in real time or NRCAN for example and / or CGGTTS files.
> >
> > I have also available for these measurements a frequency comb from menlo
> > systems. Can I use repetition rate adev to evaluate the BVA? (locking it
> in
> > a Cs beam and then in BVA + Cs beam)?
>
> If the frequency comb is CEO stabilized (aka f-2f lock) and you can
> stabilize its repetition rate using an optical cavity (10cm ULE is
> already enough), then you can get better stability than your BVA
> in the range up to 10s probably better than 100s, depending on your
> setup. If the comb is free running, then the intrinsic instability
> of the comb will be larger than the BVA.
>
> > We are working in cold atom clocks here also. A Cs fountain and another
> > kind of cold atom Cs clock. But now, we're in the construction phase. No
> > measurements being made.
>
> Oh nice! Do you already have any publications on this?
> I could only find [1] and [2].
>
> BTW: If you want some known-good reference. The Canadians (IIRC the
> NRC itself) sell caesium fountains for a reasonable price.
>
>
> > About getting the maser working again, do you have any experience with
> that
> > maser? Is the KVARZ CH1-75A.
>
> I have never personally worked on a AHM myself. I do know their
> working principle and how the electronics is supposed to look.
> Unfortunately, I don't know anyone at Kvarz, only at the former
> T4-Science.
>
>
> Attila Kinali
>
>
> [1] "Progress in Brazilian Cesium atomic fountain-BrCsF",
> by Rodriguez, Bagnato, Bueno et al, 2016
> https://doi.org/10.1109/FCS.2016.7546818
>
> [2] "Progress toward Brazilian cesium fountain second generation",
> by Bueno, Rodriguez, Müller, et al. 2018
> http://dx.doi.org/10.1088/1742-6596/975/1/012071
>
>
>
> --
> The driving force behind research is the question: "Why?"
> There are things we don't understand and things we always
> wonder about. And that's why we do research.
> -- Kobayashi Makoto
>
TV
Tom Van Baak
Fri, May 29, 2026 12:23 AM
Both Tom and I have run into that before.
> Both Tom and I have run into that before.
There's a set of photos of the HFO repair here:
http://www.leapsecond.com/museum/ch175/coil.htm
/tvb
AK
Attila Kinali
Fri, May 29, 2026 8:44 PM
I Did tought that even low noise components that I have here (resistors,
capacitor opamps (like OPA637 and OP37/27)) may contour these issues.
At first this would seem the case. But once you calculate the
contribution of the various elements needed to control the EFC in
an analog PLL system with extremely low bandwith (or large time
constants), the analog components noise and drift will dominate
the stability, especially with such superb OCXOs like the BVA.
It's really best to just ground the EFC and use a digital scheme,
either some DDS based system like the one from Rick Karlquist[1,2]
and the up-then-down conversion systems like the one formerly sold
by SKK Electronics[3], or some off-the-shelf high resolution frequency
synthesizer.
If you go for the off-the-shelf synthesizer, I would suggest you
set it to a higher frequency and then divide it down. This way
you get lower phase noise, less spurs and higher resolution.
I recommend using something akin to a lambda divider[4], as they are
very simple devices and still achieve close to theoretic performance.
We also have an optical cavity that is in progress to lock CEO. But this
will take a while I guess. The comb is repetition rate locked to BVA and
locked itself in the CEO.
Cool. So you are half way there!
Now you need to lock the frequency comb to an optical cavity.
The trick here is that an optical cavity can reach increadibly
high quality factor Q values. Much higher than anything in the
radio frequency and microwave domain could. The frequency comb
is then used to divide the optical frequency down to RF range
so it can be used in electronic measurements.
This is also the trick behind how the PTB CSF2 reaches the
short to midterm stability of an acvite hydrogen maser [5]
Have a nice weekend!
Attila Kinali
[1] "A Narrow Band High Resolution Synthesizer Using a Direct
Digital Synthesizer Followed by Repeated Dividing And Mixing",
by Rick Karlquist, 1995
[2] "A 3 to 30MHz High-Resolution Synthesizer Consisting of a
DDS, Divde-and-Mix Modules and a M/N Synthesizer", by Rick Karlquist, 1996
[3] "Characterization of an Auxiliary offset Generator for steering
of H Masers", by Rovera, Abgrall, Siccardi, 2012
[4] "The Sampling Theorem in Pi and Lambda Digital Frequency Dividers",
by Claudio Calosso and Enrico Rubiola, 2013.
[5] "Advances in the accuracy, stability, and reliability of the PTB
primary fountain clocks", by Weyers, Gerginov, Kazda, Rahm, Lipphardt,
Dobrev, and Gibble, 2018
https://doi.org/10.1088/1681-7575/aae008
--
The driving force behind research is the question: "Why?"
There are things we don't understand and things we always
wonder about. And that's why we do research.
-- Kobayashi Makoto
Ola Luiz Paulo!
On Thu, 28 May 2026 15:13:08 -0300
Luiz Paulo Damaceno via time-nuts <time-nuts@lists.febo.com> wrote:
> I Did tought that even low noise components that I have here (resistors,
> capacitor opamps (like OPA637 and OP37/27)) may contour these issues.
At first this would seem the case. But once you calculate the
contribution of the various elements needed to control the EFC in
an analog PLL system with extremely low bandwith (or large time
constants), the analog components noise and drift will dominate
the stability, especially with such superb OCXOs like the BVA.
It's really best to just ground the EFC and use a digital scheme,
either some DDS based system like the one from Rick Karlquist[1,2]
and the up-then-down conversion systems like the one formerly sold
by SKK Electronics[3], or some off-the-shelf high resolution frequency
synthesizer.
If you go for the off-the-shelf synthesizer, I would suggest you
set it to a higher frequency and then divide it down. This way
you get lower phase noise, less spurs and higher resolution.
I recommend using something akin to a lambda divider[4], as they are
very simple devices and still achieve close to theoretic performance.
> We also have an optical cavity that is in progress to lock CEO. But this
> will take a while I guess. The comb is repetition rate locked to BVA and
> locked itself in the CEO.
Cool. So you are half way there!
Now you need to lock the frequency comb to an optical cavity.
The trick here is that an optical cavity can reach increadibly
high quality factor Q values. Much higher than anything in the
radio frequency and microwave domain could. The frequency comb
is then used to divide the optical frequency down to RF range
so it can be used in electronic measurements.
This is also the trick behind how the PTB CSF2 reaches the
short to midterm stability of an acvite hydrogen maser [5]
Have a nice weekend!
Attila Kinali
[1] "A Narrow Band High Resolution Synthesizer Using a Direct
Digital Synthesizer Followed by Repeated Dividing And Mixing",
by Rick Karlquist, 1995
[2] "A 3 to 30MHz High-Resolution Synthesizer Consisting of a
DDS, Divde-and-Mix Modules and a M/N Synthesizer", by Rick Karlquist, 1996
[3] "Characterization of an Auxiliary offset Generator for steering
of H Masers", by Rovera, Abgrall, Siccardi, 2012
[4] "The Sampling Theorem in Pi and Lambda Digital Frequency Dividers",
by Claudio Calosso and Enrico Rubiola, 2013.
[5] "Advances in the accuracy, stability, and reliability of the PTB
primary fountain clocks", by Weyers, Gerginov, Kazda, Rahm, Lipphardt,
Dobrev, and Gibble, 2018
https://doi.org/10.1088/1681-7575/aae008
--
The driving force behind research is the question: "Why?"
There are things we don't understand and things we always
wonder about. And that's why we do research.
-- Kobayashi Makoto
BK
Bob kb8tq
Sat, May 30, 2026 12:26 PM
Hi
Analog PLL’s can work to a point. Just what that point is has been pretty well gone over already.
The 50 uf cap in this case it something I’d be concerted about.
No capacitor is ever ideal. In this case the “parallel resistance” is worth looking at. Dielectrics (all of them)
leak. That leakage is temperature (and maybe other factors …) dependent. This leads you off in the direction
of things like Teflon dielectrics. That gets both expensive and hard to find.
Bob
Ok. One "easy" first thing you can do is to more loosely couple
your BVA to your Cs beam standard. If your BVA is well aged,
it is likely to perform better than the Cs beam standard up
to 1000s. If you change the loop-bandwidth of your PLL such
that you get somewhere between 0.01Hz and 0.001Hz, then you
could make better use of your BVA. Of course, that's not
as easy as it sounds. If you have some phase measurement
equipment (e.g. a Timepod, but also a red pitaya would probably
do), you could use the BVA as input to the HROG, then compare
the phase of the HROG output and the Cs beam standard. It would
be then pretty easy to do the PLL loop on some PC and use the
HROG for the control/feedback part.
Thank you for the considerations. I've tunned the PLL for 0.00289Hz, today
i'm using a 1,1M ohm resistance and 50 uF capacitor. I can also reduce the
error voltage by changing the voltage in a potentiometer, just like a
"volume" control in audio systems. I expect to reduce a bit 0,25 times day
to day) to archive 1000s loop. I can only evaluate this using GNSS PPP - by
RTKLib in real time or NRCAN for example and / or CGGTTS files.
I have also available for these measurements a frequency comb from menlo
systems. Can I use repetition rate adev to evaluate the BVA? (locking it in
a Cs beam and then in BVA + Cs beam)?
Ok, so you don't have any legal requirement to keep a continuous
timescale. I guess your primary need is to provide a local realization
of the SI second that is as accurate as possible with as little noise
and uncertainty as possible?
Yes, no legal requirement, here in Brazil, only ONJR has the legal
requirement. Yes, my primary need is to provide a local realization of SI
second with little noise and uncertainty.
We are working in cold atom clocks here also. A Cs fountain and another
kind of cold atom Cs clock. But now, we're in the construction phase. No
measurements being made.
About getting the maser working again, do you have any experience with that
maser? Is the KVARZ CH1-75A.
Abraços,
Luiz Paulo
Em ter., 26 de mai. de 2026 às 11:27, Attila Kinali attila@kinali.ch
escreveu:
-> I want to achieve long term stability and also short term
I think both are important to us. I can figure out that things better
have now may need more clocks or a change of base clock.
Ok. One "easy" first thing you can do is to more loosely couple
your BVA to your Cs beam standard. If your BVA is well aged,
it is likely to perform better than the Cs beam standard up
to 1000s. If you change the loop-bandwidth of your PLL such
that you get somewhere between 0.01Hz and 0.001Hz, then you
could make better use of your BVA. Of course, that's not
as easy as it sounds. If you have some phase measurement
equipment (e.g. a Timepod, but also a red pitaya would probably
do), you could use the BVA as input to the HROG, then compare
the phase of the HROG output and the Cs beam standard. It would
be then pretty easy to do the PLL loop on some PC and use the
HROG for the control/feedback part.
-> Quite honestly, given that you are an NMI and we are talking about
a national timescale, I would recommend that you first try to enhance
your resilience against device failures. Having only a single Cs beam
standard, which need a replacement tube every 5 to 20 years (depending
on type and chance), you run the risk of losing your timescale when your
Cs beam standard runs out of Cs.
- I'm not an NMI, I'm in a university that runs atomic physics
experiments.
Oh.. sorry. My apologies! I assumed that you were an NMI as it is
rare that non-NMIs contribute to TAI/UTC.
Our UTC time scale is one way for us to have trackability to
Cs frequency realization.
- Actually we have a second Cs beam that is not running due to this:
a second one in case of failure, at least.
Ok, so you don't have any legal requirement to keep a continuous
timescale. I guess your primary need is to provide a local realization
of the SI second that is as accurate as possible with as little noise
and uncertainty as possible?
Then I would focus on getting the hydrogen maser working again
and use it as input to the HROG. That will provide you the best
short to medium term performance.
Then use your phase meter to compare your HROG output and the
Cs beam output. This way you can correct for the drift of the
maser. Use the GNSS system to correct the drift of the Cs beam
standard with a tau in the range of a few weeks to a month.
This is basically what METAS does (minus the Cs fountain that
they periodically run) [1].
BTW: If your research involves optical atomic clocks and their
absolute frequency measurement, you might want to consider getting
a primary frequency standard. In the course of the presentation on the
2025 BIPM recommended frequency list at EFTF last month, Helen Margolis
from NPL did complain about the correlations of absolute frequency
measurements done through UTC, which are very hard to properly handle.
Attila Kinali
[1] "METAS new time scale generation system - A progress report"
by Bernier, Dudle and Schlunegger, 2006
http://time.kinali.ch/ptti/2006papers/paper3.pdf
The driving force behind research is the question: "Why?"
There are things we don't understand and things we always
wonder about. And that's why we do research.
-- Kobayashi Makoto
Hi
Analog PLL’s can work to a point. Just what that point is has been pretty well gone over already.
The 50 uf cap in this case it something I’d be concerted about.
No capacitor is ever ideal. In this case the “parallel resistance” is worth looking at. Dielectrics (all of them)
leak. That leakage is temperature (and maybe other factors …) dependent. This leads you off in the direction
of things like Teflon dielectrics. That gets both expensive and hard to find.
Bob
> On May 28, 2026, at 9:37 AM, Luiz Paulo Damaceno via time-nuts <time-nuts@lists.febo.com> wrote:
>
> Ola, Attila!
>
>> Ok. One "easy" first thing you can do is to more loosely couple
>> your BVA to your Cs beam standard. If your BVA is well aged,
>> it is likely to perform better than the Cs beam standard up
>> to 1000s. If you change the loop-bandwidth of your PLL such
>> that you get somewhere between 0.01Hz and 0.001Hz, then you
>> could make better use of your BVA. Of course, that's not
>> as easy as it sounds. If you have some phase measurement
>> equipment (e.g. a Timepod, but also a red pitaya would probably
>> do), you could use the BVA as input to the HROG, then compare
>> the phase of the HROG output and the Cs beam standard. It would
>> be then pretty easy to do the PLL loop on some PC and use the
>> HROG for the control/feedback part.
>
> Thank you for the considerations. I've tunned the PLL for 0.00289Hz, today
> i'm using a 1,1M ohm resistance and 50 uF capacitor. I can also reduce the
> error voltage by changing the voltage in a potentiometer, just like a
> "volume" control in audio systems. I expect to reduce a bit 0,25 times day
> to day) to archive 1000s loop. I can only evaluate this using GNSS PPP - by
> RTKLib in real time or NRCAN for example and / or CGGTTS files.
>
> I have also available for these measurements a frequency comb from menlo
> systems. Can I use repetition rate adev to evaluate the BVA? (locking it in
> a Cs beam and then in BVA + Cs beam)?
>
>> Ok, so you don't have any legal requirement to keep a continuous
>> timescale. I guess your primary need is to provide a local realization
>> of the SI second that is as accurate as possible with as little noise
>> and uncertainty as possible?
>
> Yes, no legal requirement, here in Brazil, only ONJR has the legal
> requirement. Yes, my primary need is to provide a local realization of SI
> second with little noise and uncertainty.
>
> We are working in cold atom clocks here also. A Cs fountain and another
> kind of cold atom Cs clock. But now, we're in the construction phase. No
> measurements being made.
>
> About getting the maser working again, do you have any experience with that
> maser? Is the KVARZ CH1-75A.
>
> Abraços,
>
> Luiz Paulo
>
> Em ter., 26 de mai. de 2026 às 11:27, Attila Kinali <attila@kinali.ch>
> escreveu:
>
>> Ola Luiz Paulo!
>>
>> On Tue, 26 May 2026 10:21:03 -0300
>> Luiz Paulo Damaceno <luizpauloeletrico42@gmail.com> wrote:
>>
>>> -> I want to achieve long term stability and also short term
>> performance.
>>> I think both are important to us. I can figure out that things better
>> than we
>>> have now may need more clocks or a change of base clock.
>>
>> Ok. One "easy" first thing you can do is to more loosely couple
>> your BVA to your Cs beam standard. If your BVA is well aged,
>> it is likely to perform better than the Cs beam standard up
>> to 1000s. If you change the loop-bandwidth of your PLL such
>> that you get somewhere between 0.01Hz and 0.001Hz, then you
>> could make better use of your BVA. Of course, that's not
>> as easy as it sounds. If you have some phase measurement
>> equipment (e.g. a Timepod, but also a red pitaya would probably
>> do), you could use the BVA as input to the HROG, then compare
>> the phase of the HROG output and the Cs beam standard. It would
>> be then pretty easy to do the PLL loop on some PC and use the
>> HROG for the control/feedback part.
>>
>>> -> Quite honestly, given that you are an NMI and we are talking about
>>> a national timescale, I would recommend that you first try to enhance
>>> your resilience against device failures. Having only a single Cs beam
>>> standard, which need a replacement tube every 5 to 20 years (depending
>>> on type and chance), you run the risk of losing your timescale when your
>>> Cs beam standard runs out of Cs.
>>> - I'm not an NMI, I'm in a university that runs atomic physics
>>> experiments.
>>
>> Oh.. sorry. My apologies! I assumed that you were an NMI as it is
>> rare that non-NMIs contribute to TAI/UTC.
>>
>>> Our UTC time scale is one way for us to have trackability to
>>> Cs frequency realization.
>>> - Actually we have a second Cs beam that is not running due to this:
>> have
>>> a second one in case of failure, at least.
>>
>> Ok, so you don't have any legal requirement to keep a continuous
>> timescale. I guess your primary need is to provide a local realization
>> of the SI second that is as accurate as possible with as little noise
>> and uncertainty as possible?
>>
>> Then I would focus on getting the hydrogen maser working again
>> and use it as input to the HROG. That will provide you the best
>> short to medium term performance.
>>
>> Then use your phase meter to compare your HROG output and the
>> Cs beam output. This way you can correct for the drift of the
>> maser. Use the GNSS system to correct the drift of the Cs beam
>> standard with a tau in the range of a few weeks to a month.
>>
>> This is basically what METAS does (minus the Cs fountain that
>> they periodically run) [1].
>>
>> BTW: If your research involves optical atomic clocks and their
>> absolute frequency measurement, you might want to consider getting
>> a primary frequency standard. In the course of the presentation on the
>> 2025 BIPM recommended frequency list at EFTF last month, Helen Margolis
>> from NPL did complain about the correlations of absolute frequency
>> measurements done through UTC, which are very hard to properly handle.
>>
>>
>> Attila Kinali
>>
>>
>> [1] "METAS new time scale generation system - A progress report"
>> by Bernier, Dudle and Schlunegger, 2006
>> http://time.kinali.ch/ptti/2006papers/paper3.pdf
>> --
>> The driving force behind research is the question: "Why?"
>> There are things we don't understand and things we always
>> wonder about. And that's why we do research.
>> -- Kobayashi Makoto
>>
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com