AT
Andy Talbot
Wed, Dec 7, 2022 9:38 AM
I've been given a defunct Z3801 with a fully working ovened oscillator.
The rest of the GPSDO circuitry is just junk and will be chucked, but I
want to get the DOXCO running as a backup high stability reference here.
I've worked out the power connections for the oscillator and inner oven,
and have powered it up with +12V & +24V supplies. It's currently just
warmed up and is happily generating a nice sine at 10MHz, with +0.3V (at
the moment) t on the EFC.
I applied 24V to the outer oven earlier to see what happened, and measured
the thermistor resistance as it warmed, watching it fall from 110kΩat room
temp to 30kΩ where I stopped. I know the controller for the outer oven is
in external circuitry, but a brief web search failed to show a) what
temperature it should run at, To, and b) what the thermistor resistance
would be at that temperature, Ro.
Does anyone here know these figures off hand, or do I need to go to the
web and download the circuit and trace out the outer oven controller
details. I'll be doing my own controller circuitry, so just need To and
Ro.
Andy
www.g4jnt.com
> I've been given a defunct Z3801 with a fully working ovened oscillator.
> The rest of the GPSDO circuitry is just junk and will be chucked, but I
> want to get the DOXCO running as a backup high stability reference here.
>
> I've worked out the power connections for the oscillator and inner oven,
> and have powered it up with +12V & +24V supplies. It's currently just
> warmed up and is happily generating a nice sine at 10MHz, with +0.3V (at
> the moment) t on the EFC.
>
> I applied 24V to the outer oven earlier to see what happened, and measured
> the thermistor resistance as it warmed, watching it fall from 110kΩat room
> temp to 30kΩ where I stopped. I know the controller for the outer oven is
> in external circuitry, but a brief web search failed to show a) what
> temperature it should run at, To, and b) what the thermistor resistance
> would be at that temperature, Ro.
>
> Does anyone here know these figures off hand, or do I need to go to the
> web and download the circuit and trace out the outer oven controller
> details. I'll be doing my own controller circuitry, so just need To and
> Ro.
>
> Andy
> www.g4jnt.com
>
>
PS
paul swed
Wed, Dec 7, 2022 1:52 PM
Andy
For what you are doing I am not sure the outer ovens really needed. Its use
would have been for locations like Cel Towers and such that can be very
poorly heated if at all. I might guess you could put a temp probe in and
set it at some temp like 100 F and let it go. Or as I say not use it at all.
Others will have stronger opinions.
Regards
Paul
WB8TSL
On Wed, Dec 7, 2022 at 5:28 AM Andy Talbot via time-nuts <
time-nuts@lists.febo.com> wrote:
I've been given a defunct Z3801 with a fully working ovened oscillator.
The rest of the GPSDO circuitry is just junk and will be chucked, but I
want to get the DOXCO running as a backup high stability reference here.
I've worked out the power connections for the oscillator and inner oven,
and have powered it up with +12V & +24V supplies. It's currently just
warmed up and is happily generating a nice sine at 10MHz, with +0.3V (at
the moment) t on the EFC.
I applied 24V to the outer oven earlier to see what happened, and
the thermistor resistance as it warmed, watching it fall from 110kΩat
temp to 30kΩ where I stopped. I know the controller for the outer oven
in external circuitry, but a brief web search failed to show a) what
temperature it should run at, To, and b) what the thermistor resistance
would be at that temperature, Ro.
Does anyone here know these figures off hand, or do I need to go to the
web and download the circuit and trace out the outer oven controller
details. I'll be doing my own controller circuitry, so just need To and
Ro.
Andy
www.g4jnt.com
Andy
For what you are doing I am not sure the outer ovens really needed. Its use
would have been for locations like Cel Towers and such that can be very
poorly heated if at all. I might guess you could put a temp probe in and
set it at some temp like 100 F and let it go. Or as I say not use it at all.
Others will have stronger opinions.
Regards
Paul
WB8TSL
On Wed, Dec 7, 2022 at 5:28 AM Andy Talbot via time-nuts <
time-nuts@lists.febo.com> wrote:
> > I've been given a defunct Z3801 with a fully working ovened oscillator.
> > The rest of the GPSDO circuitry is just junk and will be chucked, but I
> > want to get the DOXCO running as a backup high stability reference here.
> >
> > I've worked out the power connections for the oscillator and inner oven,
> > and have powered it up with +12V & +24V supplies. It's currently just
> > warmed up and is happily generating a nice sine at 10MHz, with +0.3V (at
> > the moment) t on the EFC.
> >
> > I applied 24V to the outer oven earlier to see what happened, and
> measured
> > the thermistor resistance as it warmed, watching it fall from 110kΩat
> room
> > temp to 30kΩ where I stopped. I know the controller for the outer oven
> is
> > in external circuitry, but a brief web search failed to show a) what
> > temperature it should run at, To, and b) what the thermistor resistance
> > would be at that temperature, Ro.
> >
> > Does anyone here know these figures off hand, or do I need to go to the
> > web and download the circuit and trace out the outer oven controller
> > details. I'll be doing my own controller circuitry, so just need To and
> > Ro.
> >
> > Andy
> > www.g4jnt.com
> >
> >
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
BK
Bob kb8tq
Wed, Dec 7, 2022 2:07 PM
Hi
Will your lab be running below freezing? If not, just ignore the outer heater and move on.
The main reason that heater exists is a very early requirement to have the reference operate
at (and quickly warm up from) -40C. The cell guys pretty much gave up on that idea after they
dug into all of the grubby details of the rest of the gear.
Bob
I've been given a defunct Z3801 with a fully working ovened oscillator.
The rest of the GPSDO circuitry is just junk and will be chucked, but I
want to get the DOXCO running as a backup high stability reference here.
I've worked out the power connections for the oscillator and inner oven,
and have powered it up with +12V & +24V supplies. It's currently just
warmed up and is happily generating a nice sine at 10MHz, with +0.3V (at
the moment) t on the EFC.
I applied 24V to the outer oven earlier to see what happened, and measured
the thermistor resistance as it warmed, watching it fall from 110kΩat room
temp to 30kΩ where I stopped. I know the controller for the outer oven is
in external circuitry, but a brief web search failed to show a) what
temperature it should run at, To, and b) what the thermistor resistance
would be at that temperature, Ro.
Does anyone here know these figures off hand, or do I need to go to the
web and download the circuit and trace out the outer oven controller
details. I'll be doing my own controller circuitry, so just need To and
Ro.
Andy
www.g4jnt.com
Hi
Will your lab be running below freezing? If not, just ignore the outer heater and move on.
The main reason that heater exists is a very early requirement to have the reference operate
at (and quickly warm up from) -40C. The cell guys pretty much gave up on that idea after they
dug into all of the grubby details of the rest of the gear.
Bob
> On Dec 7, 2022, at 4:38 AM, Andy Talbot via time-nuts <time-nuts@lists.febo.com> wrote:
>
>> I've been given a defunct Z3801 with a fully working ovened oscillator.
>> The rest of the GPSDO circuitry is just junk and will be chucked, but I
>> want to get the DOXCO running as a backup high stability reference here.
>>
>> I've worked out the power connections for the oscillator and inner oven,
>> and have powered it up with +12V & +24V supplies. It's currently just
>> warmed up and is happily generating a nice sine at 10MHz, with +0.3V (at
>> the moment) t on the EFC.
>>
>> I applied 24V to the outer oven earlier to see what happened, and measured
>> the thermistor resistance as it warmed, watching it fall from 110kΩat room
>> temp to 30kΩ where I stopped. I know the controller for the outer oven is
>> in external circuitry, but a brief web search failed to show a) what
>> temperature it should run at, To, and b) what the thermistor resistance
>> would be at that temperature, Ro.
>>
>> Does anyone here know these figures off hand, or do I need to go to the
>> web and download the circuit and trace out the outer oven controller
>> details. I'll be doing my own controller circuitry, so just need To and
>> Ro.
>>
>> Andy
>> www.g4jnt.com
>>
>>
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
AT
Andy Talbot
Wed, Dec 7, 2022 2:12 PM
Ah, OK. It's working great on just the inner one at the moment
The temp is unlikely to drop below 15 to 16C - shack room temperature worst
case winter nights.
Andy
www.g4jnt.com
On Wed, 7 Dec 2022 at 14:07, Bob kb8tq kb8tq@n1k.org wrote:
Hi
Will your lab be running below freezing? If not, just ignore the outer
heater and move on.
The main reason that heater exists is a very early requirement to have the
reference operate
at (and quickly warm up from) -40C. The cell guys pretty much gave up on
that idea after they
dug into all of the grubby details of the rest of the gear.
Bob
On Dec 7, 2022, at 4:38 AM, Andy Talbot via time-nuts <
I've been given a defunct Z3801 with a fully working ovened oscillator.
The rest of the GPSDO circuitry is just junk and will be chucked, but I
want to get the DOXCO running as a backup high stability reference here.
I've worked out the power connections for the oscillator and inner oven,
and have powered it up with +12V & +24V supplies. It's currently just
warmed up and is happily generating a nice sine at 10MHz, with +0.3V (at
the moment) t on the EFC.
I applied 24V to the outer oven earlier to see what happened, and
the thermistor resistance as it warmed, watching it fall from 110kΩat
temp to 30kΩ where I stopped. I know the controller for the outer oven
in external circuitry, but a brief web search failed to show a) what
temperature it should run at, To, and b) what the thermistor resistance
would be at that temperature, Ro.
Does anyone here know these figures off hand, or do I need to go to the
web and download the circuit and trace out the outer oven controller
details. I'll be doing my own controller circuitry, so just need To
Ah, OK. It's working great on just the inner one at the moment
The temp is unlikely to drop below 15 to 16C - shack room temperature worst
case winter nights.
Andy
www.g4jnt.com
On Wed, 7 Dec 2022 at 14:07, Bob kb8tq <kb8tq@n1k.org> wrote:
> Hi
>
> Will your lab be running below freezing? If not, just ignore the outer
> heater and move on.
> The main reason that heater exists is a very early requirement to have the
> reference operate
> at (and quickly warm up from) -40C. The cell guys pretty much gave up on
> that idea after they
> dug into all of the grubby details of the rest of the gear.
>
> Bob
>
> > On Dec 7, 2022, at 4:38 AM, Andy Talbot via time-nuts <
> time-nuts@lists.febo.com> wrote:
> >
> >> I've been given a defunct Z3801 with a fully working ovened oscillator.
> >> The rest of the GPSDO circuitry is just junk and will be chucked, but I
> >> want to get the DOXCO running as a backup high stability reference here.
> >>
> >> I've worked out the power connections for the oscillator and inner oven,
> >> and have powered it up with +12V & +24V supplies. It's currently just
> >> warmed up and is happily generating a nice sine at 10MHz, with +0.3V (at
> >> the moment) t on the EFC.
> >>
> >> I applied 24V to the outer oven earlier to see what happened, and
> measured
> >> the thermistor resistance as it warmed, watching it fall from 110kΩat
> room
> >> temp to 30kΩ where I stopped. I know the controller for the outer oven
> is
> >> in external circuitry, but a brief web search failed to show a) what
> >> temperature it should run at, To, and b) what the thermistor resistance
> >> would be at that temperature, Ro.
> >>
> >> Does anyone here know these figures off hand, or do I need to go to the
> >> web and download the circuit and trace out the outer oven controller
> >> details. I'll be doing my own controller circuitry, so just need To
> and
> >> Ro.
> >>
> >> Andy
> >> www.g4jnt.com
> >>
> >>
> > _______________________________________________
> > time-nuts mailing list -- time-nuts@lists.febo.com
> > To unsubscribe send an email to time-nuts-leave@lists.febo.com
>
>
D
DM
Wed, Dec 7, 2022 3:27 PM
Andy,
Here's a link to a description of the outer oven and its controller on the KO4BB manuals site.
http://www.ko4bb.com/getsimple/index.php?id=manuals&dir=02_GPS_Timing/Z3801/Z3801A_Outer_Oven/Web_Page.
However, on the link to the document containing the text, all of the images have gotten lost and are not visible.
Fortunately, I had downloaded the complete page and all the files, and put them all into a PDF for my own reference. I've attached it to this message. It contains an in-depth description of the outer oven and its controller circuitry, and I think it contains all the information you're after.
Enjoy!
DaveM
----- Original Message -----
From: "Time-Nuts" time-nuts@lists.febo.com
To: "Time-Nuts" time-nuts@lists.febo.com
Cc: "Andy Talbot" andy.g4jnt@gmail.com
Sent: Wednesday, December 7, 2022 3:38:54 AM
Subject: [time-nuts] Re: DOXCO from Z3801
I've been given a defunct Z3801 with a fully working ovened oscillator.
The rest of the GPSDO circuitry is just junk and will be chucked, but I
want to get the DOXCO running as a backup high stability reference here.
I've worked out the power connections for the oscillator and inner oven,
and have powered it up with +12V & +24V supplies. It's currently just
warmed up and is happily generating a nice sine at 10MHz, with +0.3V (at
the moment) t on the EFC.
I applied 24V to the outer oven earlier to see what happened, and measured
the thermistor resistance as it warmed, watching it fall from 110kΩat room
temp to 30kΩ where I stopped. I know the controller for the outer oven is
in external circuitry, but a brief web search failed to show a) what
temperature it should run at, To, and b) what the thermistor resistance
would be at that temperature, Ro.
Does anyone here know these figures off hand, or do I need to go to the
web and download the circuit and trace out the outer oven controller
details. I'll be doing my own controller circuitry, so just need To and
Ro.
Andy
www.g4jnt.com
Andy,
Here's a link to a description of the outer oven and its controller on the KO4BB manuals site.
http://www.ko4bb.com/getsimple/index.php?id=manuals&dir=02_GPS_Timing/Z3801/Z3801A_Outer_Oven/Web_Page.
However, on the link to the document containing the text, all of the images have gotten lost and are not visible.
Fortunately, I had downloaded the complete page and all the files, and put them all into a PDF for my own reference. I've attached it to this message. It contains an in-depth description of the outer oven and its controller circuitry, and I think it contains all the information you're after.
Enjoy!
DaveM
----- Original Message -----
From: "Time-Nuts" <time-nuts@lists.febo.com>
To: "Time-Nuts" <time-nuts@lists.febo.com>
Cc: "Andy Talbot" <andy.g4jnt@gmail.com>
Sent: Wednesday, December 7, 2022 3:38:54 AM
Subject: [time-nuts] Re: DOXCO from Z3801
> I've been given a defunct Z3801 with a fully working ovened oscillator.
> The rest of the GPSDO circuitry is just junk and will be chucked, but I
> want to get the DOXCO running as a backup high stability reference here.
>
> I've worked out the power connections for the oscillator and inner oven,
> and have powered it up with +12V & +24V supplies. It's currently just
> warmed up and is happily generating a nice sine at 10MHz, with +0.3V (at
> the moment) t on the EFC.
>
> I applied 24V to the outer oven earlier to see what happened, and measured
> the thermistor resistance as it warmed, watching it fall from 110kΩat room
> temp to 30kΩ where I stopped. I know the controller for the outer oven is
> in external circuitry, but a brief web search failed to show a) what
> temperature it should run at, To, and b) what the thermistor resistance
> would be at that temperature, Ro.
>
> Does anyone here know these figures off hand, or do I need to go to the
> web and download the circuit and trace out the outer oven controller
> details. I'll be doing my own controller circuitry, so just need To and
> Ro.
>
> Andy
> www.g4jnt.com
>
>
_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
EB
ed breya
Wed, Dec 7, 2022 6:17 PM
I don't recall seeing that anyone ever reported the official temperature
setpoint for the outer oven. Most of the info on the Z3801A still
resides on realhamradio.com.
I found in my notes a paper printout of a webpage apparently put up
someone named Shegnauer, circa 2006, showing the schematic, and some
info and estimates for the temperature range, and a spice model of the
control system operation. It's a nice document. I tried to find it
online now, but it seems to have disappeared over the years. The same
link showed up in old messages on time nuts and other lists, but it
appears dead now. It must have been pretty well known, so it's likely
still floating around out there at some URL or in someone's files. If
you don't have it already, you should try to find it. I only have the
paper copy.
Anyway, it was suggested that the outer oven should run around 60-65 deg
C, while the thermistor is around 16.2 k ohms there. Personally, I
believe this is too high, and should be around 50-55 deg C or maybe a
bit more - right near the maximum ambient rating for the Z3801A,
neglecting any internal box temperature rise. If the performance specs
are relaxed toward the higher end of operating temperature range, then
the outer oven may not have to run so high. I don't recall, but think
that may be the case.
Ed
I don't recall seeing that anyone ever reported the official temperature
setpoint for the outer oven. Most of the info on the Z3801A still
resides on realhamradio.com.
I found in my notes a paper printout of a webpage apparently put up
someone named Shegnauer, circa 2006, showing the schematic, and some
info and estimates for the temperature range, and a spice model of the
control system operation. It's a nice document. I tried to find it
online now, but it seems to have disappeared over the years. The same
link showed up in old messages on time nuts and other lists, but it
appears dead now. It must have been pretty well known, so it's likely
still floating around out there at some URL or in someone's files. If
you don't have it already, you should try to find it. I only have the
paper copy.
Anyway, it was suggested that the outer oven should run around 60-65 deg
C, while the thermistor is around 16.2 k ohms there. Personally, I
believe this is too high, and should be around 50-55 deg C or maybe a
bit more - right near the maximum ambient rating for the Z3801A,
neglecting any internal box temperature rise. If the performance specs
are relaxed toward the higher end of operating temperature range, then
the outer oven may not have to run so high. I don't recall, but think
that may be the case.
Ed
EC
Edesio Costa e Silva
Wed, Dec 7, 2022 6:40 PM
I don't recall seeing that anyone ever reported the official temperature
setpoint for the outer oven. Most of the info on the Z3801A still resides on
realhamradio.com.
I found in my notes a paper printout of a webpage apparently put up someone
named Shegnauer, circa 2006, showing the schematic, and some info and
estimates for the temperature range, and a spice model of the control system
operation. It's a nice document. I tried to find it online now, but it seems
to have disappeared over the years. The same link showed up in old messages
on time nuts and other lists, but it appears dead now. It must have been
pretty well known, so it's likely still floating around out there at some
URL or in someone's files. If you don't have it already, you should try to
find it. I only have the paper copy.
Anyway, it was suggested that the outer oven should run around 60-65 deg C,
while the thermistor is around 16.2 k ohms there. Personally, I believe this
is too high, and should be around 50-55 deg C or maybe a bit more - right
near the maximum ambient rating for the Z3801A, neglecting any internal box
temperature rise. If the performance specs are relaxed toward the higher end
of operating temperature range, then the outer oven may not have to run so
high. I don't recall, but think that may be the case.
Ed
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Hi!
Found original page at
https://web.archive.org/web/20050318020038/http://homepage.sunrise.ch/mysunrise/shegnauer/Z3801A/Outer%20Oven%20Controller/z3801a_outer_oven_description.htm
but the images are missing.
Edésio
On Wed, Dec 07, 2022 at 10:17:14AM -0800, ed breya via time-nuts wrote:
> I don't recall seeing that anyone ever reported the official temperature
> setpoint for the outer oven. Most of the info on the Z3801A still resides on
> realhamradio.com.
>
> I found in my notes a paper printout of a webpage apparently put up someone
> named Shegnauer, circa 2006, showing the schematic, and some info and
> estimates for the temperature range, and a spice model of the control system
> operation. It's a nice document. I tried to find it online now, but it seems
> to have disappeared over the years. The same link showed up in old messages
> on time nuts and other lists, but it appears dead now. It must have been
> pretty well known, so it's likely still floating around out there at some
> URL or in someone's files. If you don't have it already, you should try to
> find it. I only have the paper copy.
>
> Anyway, it was suggested that the outer oven should run around 60-65 deg C,
> while the thermistor is around 16.2 k ohms there. Personally, I believe this
> is too high, and should be around 50-55 deg C or maybe a bit more - right
> near the maximum ambient rating for the Z3801A, neglecting any internal box
> temperature rise. If the performance specs are relaxed toward the higher end
> of operating temperature range, then the outer oven may not have to run so
> high. I don't recall, but think that may be the case.
>
> Ed
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
ES
Eric Scace
Wed, Dec 7, 2022 7:59 PM
Try searching the Internet Archive for that old document whose links are now dead.
Sent from my iPhone
On Dec 7, 2022, at 11:29, ed breya via time-nuts time-nuts@lists.febo.com wrote:
I don't recall seeing that anyone ever reported the official temperature setpoint for the outer oven. Most of the info on the Z3801A still resides on realhamradio.com.
I found in my notes a paper printout of a webpage apparently put up someone named Shegnauer, circa 2006, showing the schematic, and some info and estimates for the temperature range, and a spice model of the control system operation. It's a nice document. I tried to find it online now, but it seems to have disappeared over the years. The same link showed up in old messages on time nuts and other lists, but it appears dead now. It must have been pretty well known, so it's likely still floating around out there at some URL or in someone's files. If you don't have it already, you should try to find it. I only have the paper copy.
Anyway, it was suggested that the outer oven should run around 60-65 deg C, while the thermistor is around 16.2 k ohms there. Personally, I believe this is too high, and should be around 50-55 deg C or maybe a bit more - right near the maximum ambient rating for the Z3801A, neglecting any internal box temperature rise. If the performance specs are relaxed toward the higher end of operating temperature range, then the outer oven may not have to run so high. I don't recall, but think that may be the case.
Ed
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Try searching the Internet Archive for that old document whose links are now dead.
Sent from my iPhone
> On Dec 7, 2022, at 11:29, ed breya via time-nuts <time-nuts@lists.febo.com> wrote:
>
> I don't recall seeing that anyone ever reported the official temperature setpoint for the outer oven. Most of the info on the Z3801A still resides on realhamradio.com.
>
> I found in my notes a paper printout of a webpage apparently put up someone named Shegnauer, circa 2006, showing the schematic, and some info and estimates for the temperature range, and a spice model of the control system operation. It's a nice document. I tried to find it online now, but it seems to have disappeared over the years. The same link showed up in old messages on time nuts and other lists, but it appears dead now. It must have been pretty well known, so it's likely still floating around out there at some URL or in someone's files. If you don't have it already, you should try to find it. I only have the paper copy.
>
> Anyway, it was suggested that the outer oven should run around 60-65 deg C, while the thermistor is around 16.2 k ohms there. Personally, I believe this is too high, and should be around 50-55 deg C or maybe a bit more - right near the maximum ambient rating for the Z3801A, neglecting any internal box temperature rise. If the performance specs are relaxed toward the higher end of operating temperature range, then the outer oven may not have to run so high. I don't recall, but think that may be the case.
>
> Ed
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
TV
Tom Van Baak
Wed, Dec 7, 2022 8:26 PM
Several of you are having trouble. Note that the document and all 4
images are on the k04bb manual site. Just follow the URL that DaveM
provided:
http://www.ko4bb.com/getsimple/index.php?id=manuals&dir=02_GPS_Timing/Z3801/Z3801A_Outer_Oven/Web_Page
and note there are *two* entries. One is a html file and the other a
directory. In the directory are the 4 "missing" images.
----
Or, even simpler, just grab the PDF created and attached to the post by
DaveM an hour ago.
Note these files, including copies of the datasheets, are cached in
Stefan Hegnauer's time-nuts user page:
http://leapsecond.com/u/shegnauer/
----
A copy of his Z3801A power supply article survives here:
https://web.archive.org/web/20061208181458/http://homepage.sunrise.ch/mysunrise/shegnauer/Z3801A/Power%20Supply/z3801a_power_supply.htm
and the missing schematic for that was posted here:
https://febo.com/pipermail/time-nuts_lists.febo.com/2017-September/089331.html
and is also cached in the /u/shegnauer/ directory.
Hope that helps.
/tvb
On 12/7/2022 10:40 AM, Edesio Costa e Silva wrote:
> Hi!
>
> Found original page at
> https://web.archive.org/web/20050318020038/http://homepage.sunrise.ch/mysunrise/shegnauer/Z3801A/Outer%20Oven%20Controller/z3801a_outer_oven_description.htm
> but the images are missing.
>
> Edésio
>
BK
Bob kb8tq
Wed, Dec 7, 2022 10:34 PM
Hi
The idea of an unheated cell tower was indeed what they headed towards when this oven
was “designed”. Turns out there is a lot of hardware in a cell site that is very much not happy
at all when it gets below freezing. That killed the idea of the -40 degree cell tower on top of a
telephone pole dead. However it was to late for the initial version that HP came up with. They
stuck with the heater.
On a true double oven. The outer oven improves stability. It is set up along with the inner oven
and the design is optimized for both ovens. That’s not at all the case with this oven. It’s just a
warmup heater. It actually makes the stability of the device worse in most cases. Yes that sounds
crazy. For all the deep math and examples of just why this is true …. head over to Rick’s set
of papers. It’s all in there …..
Bob
On Dec 7, 2022, at 8:52 AM, paul swed via time-nuts time-nuts@lists.febo.com wrote:
Andy
For what you are doing I am not sure the outer ovens really needed. Its use
would have been for locations like Cel Towers and such that can be very
poorly heated if at all. I might guess you could put a temp probe in and
set it at some temp like 100 F and let it go. Or as I say not use it at all.
Others will have stronger opinions.
Regards
Paul
WB8TSL
On Wed, Dec 7, 2022 at 5:28 AM Andy Talbot via time-nuts <
time-nuts@lists.febo.com> wrote:
I've been given a defunct Z3801 with a fully working ovened oscillator.
The rest of the GPSDO circuitry is just junk and will be chucked, but I
want to get the DOXCO running as a backup high stability reference here.
I've worked out the power connections for the oscillator and inner oven,
and have powered it up with +12V & +24V supplies. It's currently just
warmed up and is happily generating a nice sine at 10MHz, with +0.3V (at
the moment) t on the EFC.
I applied 24V to the outer oven earlier to see what happened, and
the thermistor resistance as it warmed, watching it fall from 110kΩat
temp to 30kΩ where I stopped. I know the controller for the outer oven
in external circuitry, but a brief web search failed to show a) what
temperature it should run at, To, and b) what the thermistor resistance
would be at that temperature, Ro.
Does anyone here know these figures off hand, or do I need to go to the
web and download the circuit and trace out the outer oven controller
details. I'll be doing my own controller circuitry, so just need To and
Ro.
Andy
www.g4jnt.com
Hi
The idea of an unheated cell tower was indeed what they headed towards when this oven
was “designed”. Turns out there is a *lot* of hardware in a cell site that is very much not happy
at all when it gets below freezing. That killed the idea of the -40 degree cell tower on top of a
telephone pole dead. However it was to late for the initial version that HP came up with. They
stuck with the heater.
On a true double oven. The outer oven improves stability. It is set up along with the inner oven
and the design is optimized for both ovens. That’s not at all the case with this oven. It’s just a
warmup heater. It actually makes the stability of the device worse in most cases. Yes that sounds
crazy. For all the deep math and examples of just *why* this is true …. head over to Rick’s set
of papers. It’s all in there …..
Bob
> On Dec 7, 2022, at 8:52 AM, paul swed via time-nuts <time-nuts@lists.febo.com> wrote:
>
> Andy
> For what you are doing I am not sure the outer ovens really needed. Its use
> would have been for locations like Cel Towers and such that can be very
> poorly heated if at all. I might guess you could put a temp probe in and
> set it at some temp like 100 F and let it go. Or as I say not use it at all.
> Others will have stronger opinions.
> Regards
> Paul
> WB8TSL
>
> On Wed, Dec 7, 2022 at 5:28 AM Andy Talbot via time-nuts <
> time-nuts@lists.febo.com> wrote:
>
>>> I've been given a defunct Z3801 with a fully working ovened oscillator.
>>> The rest of the GPSDO circuitry is just junk and will be chucked, but I
>>> want to get the DOXCO running as a backup high stability reference here.
>>>
>>> I've worked out the power connections for the oscillator and inner oven,
>>> and have powered it up with +12V & +24V supplies. It's currently just
>>> warmed up and is happily generating a nice sine at 10MHz, with +0.3V (at
>>> the moment) t on the EFC.
>>>
>>> I applied 24V to the outer oven earlier to see what happened, and
>> measured
>>> the thermistor resistance as it warmed, watching it fall from 110kΩat
>> room
>>> temp to 30kΩ where I stopped. I know the controller for the outer oven
>> is
>>> in external circuitry, but a brief web search failed to show a) what
>>> temperature it should run at, To, and b) what the thermistor resistance
>>> would be at that temperature, Ro.
>>>
>>> Does anyone here know these figures off hand, or do I need to go to the
>>> web and download the circuit and trace out the outer oven controller
>>> details. I'll be doing my own controller circuitry, so just need To and
>>> Ro.
>>>
>>> Andy
>>> www.g4jnt.com
>>>
>>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@lists.febo.com
>> To unsubscribe send an email to time-nuts-leave@lists.febo.com
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
EB
ed breya
Thu, Dec 8, 2022 6:30 AM
Hi Bob,
Can you please be more specific about this? I have a Z3801A project, and
have always assumed the double-oven is a good thing. If it's actually
detrimental, then I'd like to know more. I searched a bit in this
direction, but found nothing to support this view. If it's actually bad,
then I can delete it with a pull of a connector. It's a great
simplification if true.
Reference message:
"On a true double oven. The outer oven improves stability. It is set up
along with the inner oven
and the design is optimized for both ovens. That’s not at all the case
with this oven. It’s just a
warmup heater. It actually makes the stability of the device worse in
most cases. Yes that sounds
crazy. For all the deep math and examples of just why this is true
…. head over to Rick’s set
of papers. It’s all in there ….."
Ed
Hi Bob,
Can you please be more specific about this? I have a Z3801A project, and
have always assumed the double-oven is a good thing. If it's actually
detrimental, then I'd like to know more. I searched a bit in this
direction, but found nothing to support this view. If it's actually bad,
then I can delete it with a pull of a connector. It's a great
simplification if true.
Reference message:
"On a true double oven. The outer oven improves stability. It is set up
along with the inner oven
and the design is optimized for both ovens. That’s not at all the case
with this oven. It’s just a
warmup heater. It actually makes the stability of the device worse in
most cases. Yes that sounds
crazy. For all the deep math and examples of just *why* this is true
…. head over to Rick’s set
of papers. It’s all in there ….."
Ed
BK
Bob kb8tq
Thu, Dec 8, 2022 1:27 PM
Hi
People on the list have written big long papers about just why ovens do what they do.
To get into all the details, those papers are a much better reference than some quick
messages on an forum like this.
The quick and dirty version:
There are a number of things that impact the temperature stability of an OCXO. Gradients
are a big factor. Your heat source, sensor and resonator are never all at a single microscopic
point within the device. Tear one open and that’s what you will find. Heat comes from over
here, sensor is on that side, blank in the crystal package is here and has a thermal path to
this over here.
If you take an SC and fiddle the trap so you are in “thermometer mode”, you can measure the
thermal gain to the crystal. You will find that as you do this or that the gain goes up. Eventually
it changes sign and comes back down again. The higher it gets, the more likely some very minor
thing will have an impact on the gain ( = it’s a sharp peak ).
From talking to the folks who did the Z3801, they did not have the knowledge or skills to dig into
the insides of the 10811. They also did not have a need to do so. Their task was simple. All
that was required was to put a “blanket” over the device to make it warm up from a colder
temperature. Once that warmup was done, their heater stoped doing anything useful.
Since it was not designed to improve the stability of the OCXO, leaving it turned off will
not degrade the stability.
Bob
On Dec 8, 2022, at 1:30 AM, ed breya via time-nuts time-nuts@lists.febo.com wrote:
Hi Bob,
Can you please be more specific about this? I have a Z3801A project, and have always assumed the double-oven is a good thing. If it's actually detrimental, then I'd like to know more. I searched a bit in this direction, but found nothing to support this view. If it's actually bad, then I can delete it with a pull of a connector. It's a great simplification if true.
Reference message:
"On a true double oven. The outer oven improves stability. It is set up along with the inner oven
and the design is optimized for both ovens. That’s not at all the case with this oven. It’s just a
warmup heater. It actually makes the stability of the device worse in most cases. Yes that sounds
crazy. For all the deep math and examples of just why this is true …. head over to Rick’s set
of papers. It’s all in there ….."
Ed
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Hi
People on the list have written big long papers about just why ovens do what they do.
To get into all the details, those papers are a much better reference than some quick
messages on an forum like this.
The quick and dirty version:
There are a number of things that impact the temperature stability of an OCXO. Gradients
are a big factor. Your heat source, sensor and resonator are never all at a single microscopic
point within the device. Tear one open and that’s what you will find. Heat comes from over
here, sensor is on that side, blank in the crystal package is here and has a thermal path to
this over here.
If you take an SC and fiddle the trap so you are in “thermometer mode”, you can measure the
thermal gain to the crystal. You will find that as you do this or that the gain goes up. Eventually
it changes sign and comes back down again. The higher it gets, the more likely some very minor
thing will have an impact on the gain ( = it’s a sharp peak ).
From talking to the folks who did the Z3801, they did not have the knowledge or skills to dig into
the insides of the 10811. They also did not have a need to do so. Their task was simple. All
that was required was to put a “blanket” over the device to make it warm up from a colder
temperature. Once that warmup was done, their heater stoped doing anything useful.
Since it was not designed to improve the stability of the OCXO, leaving it turned off will
not degrade the stability.
Bob
> On Dec 8, 2022, at 1:30 AM, ed breya via time-nuts <time-nuts@lists.febo.com> wrote:
>
> Hi Bob,
>
> Can you please be more specific about this? I have a Z3801A project, and have always assumed the double-oven is a good thing. If it's actually detrimental, then I'd like to know more. I searched a bit in this direction, but found nothing to support this view. If it's actually bad, then I can delete it with a pull of a connector. It's a great simplification if true.
>
> Reference message:
>
> "On a true double oven. The outer oven improves stability. It is set up along with the inner oven
> and the design is optimized for both ovens. That’s not at all the case with this oven. It’s just a
> warmup heater. It actually makes the stability of the device worse in most cases. Yes that sounds
> crazy. For all the deep math and examples of just *why* this is true …. head over to Rick’s set
> of papers. It’s all in there ….."
>
> Ed
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
EB
ed breya
Fri, Dec 9, 2022 9:24 PM
I would like to figure out for sure whether the outer oven in the Z3801A
OCXO is a benefit or not. Does anyone know of actual experiments,
studies, correspondence, hypotheses, or other documents, or personal
experience/knowledge specifically pertaining to this unit, indicating
problems with the thermal stability or uselessness of the outer oven? I
have seen all sorts of commentary on this over the years, but is it
conjecture, or fact?
I have done some searching around over the years, and recently, and
found nothing specific about it. I may not know best where or how to
look, but I think I'm doing it OK - it appears sparse so far. I'd
appreciate if anyone can post up any links to such info.
I have studied some more general papers found, like Rick's ones about
designing better OCXOs. I know that the Z3801A's OCXO is far from
perfect, and the inner and outer ovens etc are not optimally arranged,
so there may be possible weaknesses, but I'm looking for actual known or
discovered problems in actual use.
If there is a fatal flaw in the design or application of the outer oven,
then I doubt it's worth trying to improve it much - so many cycles of
taking apart and reassembly would be risky to the guts.
If it actually can't be blamed for anything, then I think it should be
beneficial to the thermal stability of the OCXO inside. I think it
should help especially in holdover mode, against temperature fluctuations.
I would think that someone with a Z3801A and the right analysis gear
could do some experiments to look for thermal instabilities
(oscillations etc due to interaction of the control loops) under at
least some typical conditions (but not exhaustive, unfortunately).
Also, it should be possible to see if it helps or not in normal and
holdover modes, with the outer oven working or disabled, while subject
to certain temperature variations. I have a Z3801A, but none of the
other stuff needed to see what's actually going on.
If someone has done some of the above tests (or similar) in the past, a
report on it is a perfect example of the kind of documents I'd like to see.
Ed
I would like to figure out for sure whether the outer oven in the Z3801A
OCXO is a benefit or not. Does anyone know of actual experiments,
studies, correspondence, hypotheses, or other documents, or personal
experience/knowledge specifically pertaining to this unit, indicating
problems with the thermal stability or uselessness of the outer oven? I
have seen all sorts of commentary on this over the years, but is it
conjecture, or fact?
I have done some searching around over the years, and recently, and
found nothing specific about it. I may not know best where or how to
look, but I think I'm doing it OK - it appears sparse so far. I'd
appreciate if anyone can post up any links to such info.
I have studied some more general papers found, like Rick's ones about
designing better OCXOs. I know that the Z3801A's OCXO is far from
perfect, and the inner and outer ovens etc are not optimally arranged,
so there may be possible weaknesses, but I'm looking for actual known or
discovered problems in actual use.
If there is a fatal flaw in the design or application of the outer oven,
then I doubt it's worth trying to improve it much - so many cycles of
taking apart and reassembly would be risky to the guts.
If it actually can't be blamed for anything, then I think it should be
beneficial to the thermal stability of the OCXO inside. I think it
should help especially in holdover mode, against temperature fluctuations.
I would think that someone with a Z3801A and the right analysis gear
could do some experiments to look for thermal instabilities
(oscillations etc due to interaction of the control loops) under at
least some typical conditions (but not exhaustive, unfortunately).
Also, it should be possible to see if it helps or not in normal and
holdover modes, with the outer oven working or disabled, while subject
to certain temperature variations. I have a Z3801A, but none of the
other stuff needed to see what's actually going on.
If someone has done some of the above tests (or similar) in the past, a
report on it is a perfect example of the kind of documents I'd like to see.
Ed
BC
Bob Camp
Fri, Dec 9, 2022 11:33 PM
Hi
Other than sitting down and talking to the folks who actually designed it and who
ran the design process, no I don’t have any direct info. I do know that, based on 40+
years as an OCXO designer that the approach used is nonsense. You could do a much
better job if stability was the target. Since it was not why they did it this way, that’s in
no way a knock on those folks. They did what they needed to do.
Bob
On Dec 9, 2022, at 4:24 PM, ed breya via time-nuts time-nuts@lists.febo.com wrote:
I would like to figure out for sure whether the outer oven in the Z3801A OCXO is a benefit or not. Does anyone know of actual experiments, studies, correspondence, hypotheses, or other documents, or personal experience/knowledge specifically pertaining to this unit, indicating problems with the thermal stability or uselessness of the outer oven? I have seen all sorts of commentary on this over the years, but is it conjecture, or fact?
I have done some searching around over the years, and recently, and found nothing specific about it. I may not know best where or how to look, but I think I'm doing it OK - it appears sparse so far. I'd appreciate if anyone can post up any links to such info.
I have studied some more general papers found, like Rick's ones about designing better OCXOs. I know that the Z3801A's OCXO is far from perfect, and the inner and outer ovens etc are not optimally arranged, so there may be possible weaknesses, but I'm looking for actual known or discovered problems in actual use.
If there is a fatal flaw in the design or application of the outer oven, then I doubt it's worth trying to improve it much - so many cycles of taking apart and reassembly would be risky to the guts.
If it actually can't be blamed for anything, then I think it should be beneficial to the thermal stability of the OCXO inside. I think it should help especially in holdover mode, against temperature fluctuations.
I would think that someone with a Z3801A and the right analysis gear could do some experiments to look for thermal instabilities (oscillations etc due to interaction of the control loops) under at least some typical conditions (but not exhaustive, unfortunately).
Also, it should be possible to see if it helps or not in normal and holdover modes, with the outer oven working or disabled, while subject to certain temperature variations. I have a Z3801A, but none of the other stuff needed to see what's actually going on.
If someone has done some of the above tests (or similar) in the past, a report on it is a perfect example of the kind of documents I'd like to see.
Ed
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Hi
Other than sitting down and talking to the folks who actually designed it and who
ran the design process, no I don’t have any direct info. I do know that, based on 40+
years as an OCXO designer that the approach used is nonsense. You could do a much
better job *if* stability was the target. Since it was not why they did it this way, that’s in
no way a knock on those folks. They did what they needed to do.
Bob
> On Dec 9, 2022, at 4:24 PM, ed breya via time-nuts <time-nuts@lists.febo.com> wrote:
>
> I would like to figure out for sure whether the outer oven in the Z3801A OCXO is a benefit or not. Does anyone know of actual experiments, studies, correspondence, hypotheses, or other documents, or personal experience/knowledge specifically pertaining to this unit, indicating problems with the thermal stability or uselessness of the outer oven? I have seen all sorts of commentary on this over the years, but is it conjecture, or fact?
>
> I have done some searching around over the years, and recently, and found nothing specific about it. I may not know best where or how to look, but I think I'm doing it OK - it appears sparse so far. I'd appreciate if anyone can post up any links to such info.
>
> I have studied some more general papers found, like Rick's ones about designing better OCXOs. I know that the Z3801A's OCXO is far from perfect, and the inner and outer ovens etc are not optimally arranged, so there may be possible weaknesses, but I'm looking for actual known or discovered problems in actual use.
>
> If there is a fatal flaw in the design or application of the outer oven, then I doubt it's worth trying to improve it much - so many cycles of taking apart and reassembly would be risky to the guts.
>
> If it actually can't be blamed for anything, then I think it should be beneficial to the thermal stability of the OCXO inside. I think it should help especially in holdover mode, against temperature fluctuations.
>
> I would think that someone with a Z3801A and the right analysis gear could do some experiments to look for thermal instabilities (oscillations etc due to interaction of the control loops) under at least some typical conditions (but not exhaustive, unfortunately).
>
> Also, it should be possible to see if it helps or not in normal and holdover modes, with the outer oven working or disabled, while subject to certain temperature variations. I have a Z3801A, but none of the other stuff needed to see what's actually going on.
>
> If someone has done some of the above tests (or similar) in the past, a report on it is a perfect example of the kind of documents I'd like to see.
>
> Ed
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
SH
stefan.hegnauer (at) gmx.ch
Sun, Dec 11, 2022 9:07 PM
Ok, late to the party … lurking around from time to time.
Note that these same web pages have a new home at Stefan Hegnauer's Homepage - Electronics (selfhost.eu) . No new or additional content to what is archived at the links above, though (and thanks for archiving it in the first place!). However, links should point to sensible content.
Yes, I’m the original author of this, long time ago. Feel free to contact me if you have any questions.
-Stefan
Ok, late to the party … lurking around from time to time.
Note that these same web pages have a new home at [Stefan Hegnauer's Homepage - Electronics (selfhost.eu)](https://hegnauer.selfhost.eu/web/electronics/e_index.html) . No new or additional content to what is archived at the links above, though (and thanks for archiving it in the first place!). However, links should point to sensible content.
Yes, I’m the original author of this, long time ago. Feel free to contact me if you have any questions.
\-Stefan
EB
ed breya
Tue, Jan 17, 2023 8:37 PM
Hi Bob,
I think what may be missing is the unknown thermal resistance of the
outermost insulation surface to the ambient air. This can be very messy,
but could be estimated experimentally on an actual thing, including the
conduction, radiation, and convection effects. If there's a draft in the
room, then it's a different deal.
Deleting the foam, say, takes out its 0.04 deg C/W of resistance, but
doesn't short it to ambient temperature. If it were in a water bath or
heavy metal outer box held at Ta, it would be a different story, but
here the air interface is the big part.
Ed
Hi Bob,
I think what may be missing is the unknown thermal resistance of the
outermost insulation surface to the ambient air. This can be very messy,
but could be estimated experimentally on an actual thing, including the
conduction, radiation, and convection effects. If there's a draft in the
room, then it's a different deal.
Deleting the foam, say, takes out its 0.04 deg C/W of resistance, but
doesn't short it to ambient temperature. If it were in a water bath or
heavy metal outer box held at Ta, it would be a different story, but
here the air interface is the big part.
Ed
BC
Bob Camp
Tue, Jan 17, 2023 9:48 PM
Hi
The 4C rise with the box in place is a measure using a real box. Simply put, the stuff inside
runs at about 4C higher with the box in place relative to the room temperature than it does
with the box missing. That’s after letting things sit for many hours …. (and checking after
a few days). Given all the variables, yes, it could be 5 or 6C. It’s certainly not anything less
than 4C. The room cycles so picking a target is a bit wonky.
The box deliberately has a somewhat open top, to keep some level of convection cooling
what’s inside. Yes, that’s another un-mentioned detail and it certainly has an impact. Lots
of details like that would get involved in a precise answer to the question. One could get
all sorts of cool modeling software involved.
Right now, all I’m really looking for is an order of magnitude sort of answer. Put it in terms
of whatever your favorite material is:
How many liters or kg of this or that will it take to do the job of knocking the swing down to
under 1C over 24 hours?
That’s with the room swinging 2 to 4C over the same period.
The limit on insulation is pretty obvious: Put to much on and the device(s) inside the box
overheat and may be damaged. Some significant margin between max ambient on the
device(s) and box temp needs to be maintained.
The limitation on the thermal mass really is however much room you happen to have.
Heading down the mass route without a rough idea of what’s going to be needed does
not sound like a lot of fun. In my case a redesign of several things comes in each time
you adjust dimensions.
Obvious alternatives get you right back to some sort of servo setup running this or that
as a control element. (unless there is something else ….).
Most of the “big guys” seem to have found a convenient cave (or built one). Then then
get to publish papers on the shortcomings of thermal control in their custom built
cavern ….. All of the things they complain about apply here. The power load in the
area is variable. People (and dogs) wander through on a random basis. There’s a door
to an unheated space here and another one over there. They get opened from time
to time …. (none of that goes into the 4C number, but probably should ….. )
Bob
On Jan 17, 2023, at 3:37 PM, ed breya via time-nuts time-nuts@lists.febo.com wrote:
Hi Bob,
I think what may be missing is the unknown thermal resistance of the outermost insulation surface to the ambient air. This can be very messy, but could be estimated experimentally on an actual thing, including the conduction, radiation, and convection effects. If there's a draft in the room, then it's a different deal.
Deleting the foam, say, takes out its 0.04 deg C/W of resistance, but doesn't short it to ambient temperature. If it were in a water bath or heavy metal outer box held at Ta, it would be a different story, but here the air interface is the big part.
Ed
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Hi
The 4C rise with the box in place is a measure using a real box. Simply put, the stuff inside
runs at about 4C higher with the box in place relative to the room temperature than it does
with the box missing. That’s after letting things sit for many hours …. (and checking after
a few days). Given all the variables, yes, it could be 5 or 6C. It’s certainly not anything less
than 4C. The room cycles so picking a target is a bit wonky.
The box deliberately has a somewhat open top, to keep some level of convection cooling
what’s inside. Yes, that’s another un-mentioned detail and it certainly has an impact. Lots
of details like that would get involved in a precise answer to the question. One could get
all sorts of cool modeling software involved.
Right now, all I’m really looking for is an order of magnitude sort of answer. Put it in terms
of whatever your favorite material is:
How many liters or kg of this or that will it take to do the job of knocking the swing down to
under 1C over 24 hours?
That’s with the room swinging 2 to 4C over the same period.
The limit on insulation is pretty obvious: Put to much on and the device(s) inside the box
overheat and may be damaged. Some significant margin between max ambient on the
device(s) and box temp needs to be maintained.
The limitation on the thermal mass really is however much room you happen to have.
Heading down the mass route without a rough idea of what’s going to be needed does
not sound like a lot of fun. In my case a redesign of several things comes in each time
you adjust dimensions.
Obvious alternatives get you right back to some sort of servo setup running this or that
as a control element. (unless there is something else ….).
Most of the “big guys” seem to have found a convenient cave (or built one). Then then
get to publish papers on the shortcomings of thermal control in their custom built
cavern ….. All of the things they complain about apply here. The power load in the
area is variable. People (and dogs) wander through on a random basis. There’s a door
to an unheated space here and another one over there. They get opened from time
to time …. (none of that goes into the 4C number, but probably should ….. )
Bob
> On Jan 17, 2023, at 3:37 PM, ed breya via time-nuts <time-nuts@lists.febo.com> wrote:
>
> Hi Bob,
>
> I think what may be missing is the unknown thermal resistance of the outermost insulation surface to the ambient air. This can be very messy, but could be estimated experimentally on an actual thing, including the conduction, radiation, and convection effects. If there's a draft in the room, then it's a different deal.
>
> Deleting the foam, say, takes out its 0.04 deg C/W of resistance, but doesn't short it to ambient temperature. If it were in a water bath or heavy metal outer box held at Ta, it would be a different story, but here the air interface is the big part.
>
> Ed
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
PK
Poul-Henning Kamp
Tue, Jan 17, 2023 10:43 PM
ed breya via time-nuts writes:
I think what may be missing is the unknown thermal resistance of the
outermost insulation surface to the ambient air. This can be very messy,
but could be estimated experimentally on an actual thing, [...]
Unless you have cranked your fan or A/C up to gale force winds, you can safely ignore that inside a building,
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
--------
ed breya via time-nuts writes:
> I think what may be missing is the unknown thermal resistance of the
> outermost insulation surface to the ambient air. This can be very messy,
> but could be estimated experimentally on an actual thing, [...]
Unless you have cranked your fan or A/C up to gale force winds, you can safely ignore that inside a building,
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
NM
Neville Michie
Wed, Jan 18, 2023 4:34 AM
Watching this discussion of laboratory temperature control, I feel that my experience might be helpful. I inherited a laboratory, about 1980, about 4 X 5 X 7 metres in dimension. I had the help of a refrigeration/AC tradesman, and together we wanted to try to improve conditions.
There was a ~ 1 HP reverse cycle window conditioner in the middle of the long wall.
Measurements with a couple of thermohydrographs (clockwork recording temperature/humidity recorder)
showed swings of 6 or 8 degrees C in the course of a few hours. The temperature was controlled by a mechanical switch which had hysteresis of about 2 to 4 Degrees C. The unit operated like this: for about an hour the compressor laboured away cooling the room until the return air to the conditioner
had fallen by 2 or 3 degrees.
Then the unit would switch itself off for a few minutes (to decompress) and then it would chug away for another quarter hour it would heating the air until the room had heated 2 or 4 degrees.
More of an oscillator than a controller.
After some experiments this is what we finished up with.
A small computer fan was fixed to the wall blowing air parallel to the wall. This caused the air to slowly rotate around the room. Air velocity was less that 0.5m/s, which is the threshold to be able to sense air flow against the face. So there was no apparent airflow unless you looked for it.
The air took about a minute to travel around the room. The stirring was enough to make the air behave like a well stirred vessel. Now the room contained about 140 Kg of air, so the the heating by one kilowatt was about 0.007 degrees/second.
The air temperature was measured with a tiny glass encapsulated thermistor which was near enough to the fan to responds in one or two seconds. It turned out to quite accurately measure the average temperature of the room.
The output of the thermistor circuit was simply over or under the setpoint.
The compressor was switched with zero voltage switching industrial solid state relays.
What made this system very successful was the logic connecting the temperature sensor to the compressor, and heat/cool relay.
There were several rules to implement.
If the room was in heating mode;
if the temperature rose 0.1 degree above setpoint, the
compressor was stoped. If the compressor is ever stopped it must stay stopped for one minute.
At the end of the minute, if temperature is below the setpoint, the compressor is switched on.
For the next 5 minutes if the temperature falls below setpoint, the compressor starts. If after only 2 seconds the temperature rises above the setpoint the compressor is switched off. This is not ”short cycling” because the compressor stays off for for a minute.
Cool cycle;
If after 5 minutes of no heating demand the cooling mode is implemented.
The window unit is put into refrigeration mode, the compressor is started and keeps running until the temperature falls below the setpoint. When that is achieved, the compressor cuts out and back in, always allowing a minute(or two) for the gas pressure in the compressor to subside so it does not have to start under load. If no cooling is required for 5 or ten minutes the mode is changed.
The temperature control was miraculous. The temperature did not deviate by more that a small fraction of a degree for weeks on end.
The changeover period, from heat to cool and back, did not cause noticeable transients because they only occurred when conditions outside had cooled or heated to nearly match the setpoint of the room.
5 or 6 of these rooms were built, a paper was published, but it was attacked with papers saying it could not work by people who never tried it, and would not even come to witness it.
So stay away from 3 term controllers, they are not relevant, and commercial “Thermostats”
and use this simple on off controller either ramping up or down at 0.5 degree per minute
and with one second response time trimming swings to a few hundredths of a degree Celsius.
cheers, Neville Michie
On 18 Jan 2023, at 08:48, Bob Camp via time-nuts time-nuts@lists.febo.com wrote:
Hi
The 4C rise with the box in place is a measure using a real box. Simply put, the stuff inside
runs at about 4C higher with the box in place relative to the room temperature than it does
with the box missing. That’s after letting things sit for many hours …. (and checking after
a few days). Given all the variables, yes, it could be 5 or 6C. It’s certainly not anything less
than 4C. The room cycles so picking a target is a bit wonky.
The box deliberately has a somewhat open top, to keep some level of convection cooling
what’s inside. Yes, that’s another un-mentioned detail and it certainly has an impact. Lots
of details like that would get involved in a precise answer to the question. One could get
all sorts of cool modeling software involved.
Right now, all I’m really looking for is an order of magnitude sort of answer. Put it in terms
of whatever your favorite material is:
How many liters or kg of this or that will it take to do the job of knocking the swing down to
under 1C over 24 hours?
That’s with the room swinging 2 to 4C over the same period.
The limit on insulation is pretty obvious: Put to much on and the device(s) inside the box
overheat and may be damaged. Some significant margin between max ambient on the
device(s) and box temp needs to be maintained.
The limitation on the thermal mass really is however much room you happen to have.
Heading down the mass route without a rough idea of what’s going to be needed does
not sound like a lot of fun. In my case a redesign of several things comes in each time
you adjust dimensions.
Obvious alternatives get you right back to some sort of servo setup running this or that
as a control element. (unless there is something else ….).
Most of the “big guys” seem to have found a convenient cave (or built one). Then then
get to publish papers on the shortcomings of thermal control in their custom built
cavern ….. All of the things they complain about apply here. The power load in the
area is variable. People (and dogs) wander through on a random basis. There’s a door
to an unheated space here and another one over there. They get opened from time
to time …. (none of that goes into the 4C number, but probably should ….. )
Bob
On Jan 17, 2023, at 3:37 PM, ed breya via time-nuts time-nuts@lists.febo.com wrote:
Hi Bob,
I think what may be missing is the unknown thermal resistance of the outermost insulation surface to the ambient air. This can be very messy, but could be estimated experimentally on an actual thing, including the conduction, radiation, and convection effects. If there's a draft in the room, then it's a different deal.
Deleting the foam, say, takes out its 0.04 deg C/W of resistance, but doesn't short it to ambient temperature. If it were in a water bath or heavy metal outer box held at Ta, it would be a different story, but here the air interface is the big part.
Ed
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Watching this discussion of laboratory temperature control, I feel that my experience might be helpful. I inherited a laboratory, about 1980, about 4 X 5 X 7 metres in dimension. I had the help of a refrigeration/AC tradesman, and together we wanted to try to improve conditions.
There was a ~ 1 HP reverse cycle window conditioner in the middle of the long wall.
Measurements with a couple of thermohydrographs (clockwork recording temperature/humidity recorder)
showed swings of 6 or 8 degrees C in the course of a few hours. The temperature was controlled by a mechanical switch which had hysteresis of about 2 to 4 Degrees C. The unit operated like this: for about an hour the compressor laboured away cooling the room until the return air to the conditioner
had fallen by 2 or 3 degrees.
Then the unit would switch itself off for a few minutes (to decompress) and then it would chug away for another quarter hour it would heating the air until the room had heated 2 or 4 degrees.
More of an oscillator than a controller.
After some experiments this is what we finished up with.
A small computer fan was fixed to the wall blowing air parallel to the wall. This caused the air to slowly rotate around the room. Air velocity was less that 0.5m/s, which is the threshold to be able to sense air flow against the face. So there was no apparent airflow unless you looked for it.
The air took about a minute to travel around the room. The stirring was enough to make the air behave like a well stirred vessel. Now the room contained about 140 Kg of air, so the the heating by one kilowatt was about 0.007 degrees/second.
The air temperature was measured with a tiny glass encapsulated thermistor which was near enough to the fan to responds in one or two seconds. It turned out to quite accurately measure the average temperature of the room.
The output of the thermistor circuit was simply over or under the setpoint.
The compressor was switched with zero voltage switching industrial solid state relays.
What made this system very successful was the logic connecting the temperature sensor to the compressor, and heat/cool relay.
There were several rules to implement.
If the room was in heating mode;
if the temperature rose 0.1 degree above setpoint, the
compressor was stoped. If the compressor is ever stopped it must stay stopped for one minute.
At the end of the minute, if temperature is below the setpoint, the compressor is switched on.
For the next 5 minutes if the temperature falls below setpoint, the compressor starts. If after only 2 seconds the temperature rises above the setpoint the compressor is switched off. This is not ”short cycling” because the compressor stays off for for a minute.
Cool cycle;
If after 5 minutes of no heating demand the cooling mode is implemented.
The window unit is put into refrigeration mode, the compressor is started and keeps running until the temperature falls below the setpoint. When that is achieved, the compressor cuts out and back in, always allowing a minute(or two) for the gas pressure in the compressor to subside so it does not have to start under load. If no cooling is required for 5 or ten minutes the mode is changed.
The temperature control was miraculous. The temperature did not deviate by more that a small fraction of a degree for weeks on end.
The changeover period, from heat to cool and back, did not cause noticeable transients because they only occurred when conditions outside had cooled or heated to nearly match the setpoint of the room.
5 or 6 of these rooms were built, a paper was published, but it was attacked with papers saying it could not work by people who never tried it, and would not even come to witness it.
So stay away from 3 term controllers, they are not relevant, and commercial “Thermostats”
and use this simple on off controller either ramping up or down at 0.5 degree per minute
and with one second response time trimming swings to a few hundredths of a degree Celsius.
cheers, Neville Michie
> On 18 Jan 2023, at 08:48, Bob Camp via time-nuts <time-nuts@lists.febo.com> wrote:
>
> Hi
>
> The 4C rise with the box in place is a measure using a real box. Simply put, the stuff inside
> runs at about 4C higher with the box in place relative to the room temperature than it does
> with the box missing. That’s after letting things sit for many hours …. (and checking after
> a few days). Given all the variables, yes, it could be 5 or 6C. It’s certainly not anything less
> than 4C. The room cycles so picking a target is a bit wonky.
>
> The box deliberately has a somewhat open top, to keep some level of convection cooling
> what’s inside. Yes, that’s another un-mentioned detail and it certainly has an impact. Lots
> of details like that would get involved in a precise answer to the question. One could get
> all sorts of cool modeling software involved.
>
> Right now, all I’m really looking for is an order of magnitude sort of answer. Put it in terms
> of whatever your favorite material is:
>
> How many liters or kg of this or that will it take to do the job of knocking the swing down to
> under 1C over 24 hours?
>
> That’s with the room swinging 2 to 4C over the same period.
>
> The limit on insulation is pretty obvious: Put to much on and the device(s) inside the box
> overheat and may be damaged. Some significant margin between max ambient on the
> device(s) and box temp needs to be maintained.
>
> The limitation on the thermal mass really is however much room you happen to have.
> Heading down the mass route without a rough idea of what’s going to be needed does
> not sound like a lot of fun. In my case a redesign of several things comes in each time
> you adjust dimensions.
>
> Obvious alternatives get you right back to some sort of servo setup running this or that
> as a control element. (unless there is something else ….).
>
> Most of the “big guys” seem to have found a convenient cave (or built one). Then then
> get to publish papers on the shortcomings of thermal control in their custom built
> cavern ….. All of the things they complain about apply here. The power load in the
> area is variable. People (and dogs) wander through on a random basis. There’s a door
> to an unheated space here and another one over there. They get opened from time
> to time …. (none of that goes into the 4C number, but probably should ….. )
>
> Bob
>
>> On Jan 17, 2023, at 3:37 PM, ed breya via time-nuts <time-nuts@lists.febo.com> wrote:
>>
>> Hi Bob,
>>
>> I think what may be missing is the unknown thermal resistance of the outermost insulation surface to the ambient air. This can be very messy, but could be estimated experimentally on an actual thing, including the conduction, radiation, and convection effects. If there's a draft in the room, then it's a different deal.
>>
>> Deleting the foam, say, takes out its 0.04 deg C/W of resistance, but doesn't short it to ambient temperature. If it were in a water bath or heavy metal outer box held at Ta, it would be a different story, but here the air interface is the big part.
>>
>> Ed
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@lists.febo.com
>> To unsubscribe send an email to time-nuts-leave@lists.febo.com
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
EB
ed breya
Tue, Jan 31, 2023 8:20 PM
These midget flanged base style bulbs are extensively used in old gear.
I used to have a pretty big collection of salvaged old indicator lamp
fixtures and bulbs, but gradually used them up to now having only a few
of the really nice ones left.
I have done a lot of LED swap-ins with these, to get virtually permanent
bulb life in projects that I build. I gut the glass bulb and adhesive
out of the base, and mount a suitable LED, and design or modify the
electronics to work with the LED. The LEDs never look quite like the
original lamps - sometimes worse, sometimes better - so it usually takes
some fooling around with the jewel and LED colors and added diffusing
materials to be satisfactory.
Doing this for existing equipment almost always needs some internal
mods, which are typically quite easy. I have a fair number of certain
bulbs still good, saved for replacements in various gear. Once they run
out, it will be time for changing to LEDs. I save all the burned out
bulbs as husks for LED transplants.
Here's an interesting aside (I think). The Tek 7000 series scopes use
this style bulb (three of them) for graticule illumination. A few years
ago, I made and installed some "white" LED replacements to see how they
worked. It looked pretty good, and I was happy. Then I took some screen
shots with a phone/camera. The graticules showed up nearly invisible in
the pictures. The false spectrum of the LEDs fooled the eye OK, but the
camera's response was way different.
The lesson is to make sure the LED "color" actually works properly in
all of its applications. The LEDs I used back then were of some whitish
unknown type. I think the more modern high power ones built for lighting
should have a good spectrum for cameras too, but I haven't yet tried the
obvious experiment to find out.
Ed
These midget flanged base style bulbs are extensively used in old gear.
I used to have a pretty big collection of salvaged old indicator lamp
fixtures and bulbs, but gradually used them up to now having only a few
of the really nice ones left.
I have done a lot of LED swap-ins with these, to get virtually permanent
bulb life in projects that I build. I gut the glass bulb and adhesive
out of the base, and mount a suitable LED, and design or modify the
electronics to work with the LED. The LEDs never look quite like the
original lamps - sometimes worse, sometimes better - so it usually takes
some fooling around with the jewel and LED colors and added diffusing
materials to be satisfactory.
Doing this for existing equipment almost always needs some internal
mods, which are typically quite easy. I have a fair number of certain
bulbs still good, saved for replacements in various gear. Once they run
out, it will be time for changing to LEDs. I save all the burned out
bulbs as husks for LED transplants.
Here's an interesting aside (I think). The Tek 7000 series scopes use
this style bulb (three of them) for graticule illumination. A few years
ago, I made and installed some "white" LED replacements to see how they
worked. It looked pretty good, and I was happy. Then I took some screen
shots with a phone/camera. The graticules showed up nearly invisible in
the pictures. The false spectrum of the LEDs fooled the eye OK, but the
camera's response was way different.
The lesson is to make sure the LED "color" actually works properly in
all of its applications. The LEDs I used back then were of some whitish
unknown type. I think the more modern high power ones built for lighting
should have a good spectrum for cameras too, but I haven't yet tried the
obvious experiment to find out.
Ed
EB
ed breya
Tue, Mar 7, 2023 4:50 PM
Maybe the clock option is working just fine, but not applied properly.
It could be that the driver output is an open-collector, meant to pull a
stepper coil or such down to ground for each pulse. Without a load,
you'd see just some small collector voltage due to the base current in
the output transistor - the forward base-emitter voltage, minus the
forward base-collector junction voltage could be in the tens of mV into
high a impedance (like a scope) measurement.
So, try adding a pull-up resistor to draw at least a few mA, and see if
it looks different. Good luck.
Ed
Maybe the clock option is working just fine, but not applied properly.
It could be that the driver output is an open-collector, meant to pull a
stepper coil or such down to ground for each pulse. Without a load,
you'd see just some small collector voltage due to the base current in
the output transistor - the forward base-emitter voltage, minus the
forward base-collector junction voltage could be in the tens of mV into
high a impedance (like a scope) measurement.
So, try adding a pull-up resistor to draw at least a few mA, and see if
it looks different. Good luck.
Ed
EB
ed breya
Fri, Mar 24, 2023 8:22 PM
As always on this subject type, I must comment on the utility of
managing ground loops with common-mode chokes on cables between
equipment. I think this is the simplest, most beneficial thing you can
do for good quality signal distribution.
Good cables, shielding, and grounding are all fine and good, but do not
address the ground loop currents that inevitably flow between any
interconnected pieces of gear. Adding CM chokes raises the common-mode
impedance of the cable(s), greatly reducing the loop currents. You can
see these parts all the time, inside equipment, or added or built into
cable assemblies and such - almost always to fix EMC issues. The
simplest are the clip-on type ferrites that are easily added to almost
any cable type.
These are mainly to suppress RF and HF emissions from SMPSs and digital
circuits and such, but they can also be effective even for mains
frequency (and harmonics) ground loop control, with the right cores and
some simple tricks. The reason is that the voltage levels and impedances
are quite low (millivolts and milliohms), so adding even a small
inductive reactance can be quite effective, percentage-wise.
If you have mains frequency ground loop problems, you can try cores with
as high an A sub L as possible, and even better, multiple turns to get
the squaring benefit. There are ferrites available that should do well,
depending on the situation. I have even used regular silicon steel wound
or laminated line frequency transformer cores as CM chokes in certain
situations where very large line rejection was needed.
For narrow-band distribution of 10 MHz for instance, another option is
to use transformer coupling of the signal, so the low frequency ground
current and interference is eliminated, and a high frequency CM choke
can take care of the rest.
Ed
As always on this subject type, I must comment on the utility of
managing ground loops with common-mode chokes on cables between
equipment. I think this is the simplest, most beneficial thing you can
do for good quality signal distribution.
Good cables, shielding, and grounding are all fine and good, but do not
address the ground loop currents that inevitably flow between any
interconnected pieces of gear. Adding CM chokes raises the common-mode
impedance of the cable(s), greatly reducing the loop currents. You can
see these parts all the time, inside equipment, or added or built into
cable assemblies and such - almost always to fix EMC issues. The
simplest are the clip-on type ferrites that are easily added to almost
any cable type.
These are mainly to suppress RF and HF emissions from SMPSs and digital
circuits and such, but they can also be effective even for mains
frequency (and harmonics) ground loop control, with the right cores and
some simple tricks. The reason is that the voltage levels and impedances
are quite low (millivolts and milliohms), so adding even a small
inductive reactance can be quite effective, percentage-wise.
If you have mains frequency ground loop problems, you can try cores with
as high an A sub L as possible, and even better, multiple turns to get
the squaring benefit. There are ferrites available that should do well,
depending on the situation. I have even used regular silicon steel wound
or laminated line frequency transformer cores as CM chokes in certain
situations where very large line rejection was needed.
For narrow-band distribution of 10 MHz for instance, another option is
to use transformer coupling of the signal, so the low frequency ground
current and interference is eliminated, and a high frequency CM choke
can take care of the rest.
Ed
DB
David Bengtson
Sat, Mar 25, 2023 7:32 PM
I've come across this document
audiosystemsgroup.com/RFI-Ham.pdf
which is a good compilation of RFI minimization focused at Ham
operators, so focused at HF frequencies (Sub 30 MHz) so pretty
relevant.
Dave
On Sat, Mar 25, 2023 at 5:04 AM ed breya via time-nuts
time-nuts@lists.febo.com wrote:
As always on this subject type, I must comment on the utility of
managing ground loops with common-mode chokes on cables between
equipment. I think this is the simplest, most beneficial thing you can
do for good quality signal distribution.
Good cables, shielding, and grounding are all fine and good, but do not
address the ground loop currents that inevitably flow between any
interconnected pieces of gear. Adding CM chokes raises the common-mode
impedance of the cable(s), greatly reducing the loop currents. You can
see these parts all the time, inside equipment, or added or built into
cable assemblies and such - almost always to fix EMC issues. The
simplest are the clip-on type ferrites that are easily added to almost
any cable type.
These are mainly to suppress RF and HF emissions from SMPSs and digital
circuits and such, but they can also be effective even for mains
frequency (and harmonics) ground loop control, with the right cores and
some simple tricks. The reason is that the voltage levels and impedances
are quite low (millivolts and milliohms), so adding even a small
inductive reactance can be quite effective, percentage-wise.
If you have mains frequency ground loop problems, you can try cores with
as high an A sub L as possible, and even better, multiple turns to get
the squaring benefit. There are ferrites available that should do well,
depending on the situation. I have even used regular silicon steel wound
or laminated line frequency transformer cores as CM chokes in certain
situations where very large line rejection was needed.
For narrow-band distribution of 10 MHz for instance, another option is
to use transformer coupling of the signal, so the low frequency ground
current and interference is eliminated, and a high frequency CM choke
can take care of the rest.
Ed
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
I've come across this document
audiosystemsgroup.com/RFI-Ham.pdf
which is a good compilation of RFI minimization focused at Ham
operators, so focused at HF frequencies (Sub 30 MHz) so pretty
relevant.
Dave
On Sat, Mar 25, 2023 at 5:04 AM ed breya via time-nuts
<time-nuts@lists.febo.com> wrote:
>
> As always on this subject type, I must comment on the utility of
> managing ground loops with common-mode chokes on cables between
> equipment. I think this is the simplest, most beneficial thing you can
> do for good quality signal distribution.
>
> Good cables, shielding, and grounding are all fine and good, but do not
> address the ground loop currents that inevitably flow between any
> interconnected pieces of gear. Adding CM chokes raises the common-mode
> impedance of the cable(s), greatly reducing the loop currents. You can
> see these parts all the time, inside equipment, or added or built into
> cable assemblies and such - almost always to fix EMC issues. The
> simplest are the clip-on type ferrites that are easily added to almost
> any cable type.
>
> These are mainly to suppress RF and HF emissions from SMPSs and digital
> circuits and such, but they can also be effective even for mains
> frequency (and harmonics) ground loop control, with the right cores and
> some simple tricks. The reason is that the voltage levels and impedances
> are quite low (millivolts and milliohms), so adding even a small
> inductive reactance can be quite effective, percentage-wise.
>
> If you have mains frequency ground loop problems, you can try cores with
> as high an A sub L as possible, and even better, multiple turns to get
> the squaring benefit. There are ferrites available that should do well,
> depending on the situation. I have even used regular silicon steel wound
> or laminated line frequency transformer cores as CM chokes in certain
> situations where very large line rejection was needed.
>
> For narrow-band distribution of 10 MHz for instance, another option is
> to use transformer coupling of the signal, so the low frequency ground
> current and interference is eliminated, and a high frequency CM choke
> can take care of the rest.
>
> Ed
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
EB
ed breya
Tue, Mar 28, 2023 6:01 AM
A cold oven is nearly the only and best explanation for that much error.
The good news is that these can be readily opened up and worked on.
There's plenty of info out there.
Ed
A cold oven is nearly the only and best explanation for that much error.
The good news is that these can be readily opened up and worked on.
There's plenty of info out there.
Ed
EB
ed breya
Thu, Mar 30, 2023 4:51 PM
Wow - that is scary-burnt. Maybe it's a good idea to have some form of
thermal cutout, even if a nuisance sometimes.
Ed
Wow - that is scary-burnt. Maybe it's a good idea to have some form of
thermal cutout, even if a nuisance sometimes.
Ed
BC
Bob Camp
Thu, Mar 30, 2023 7:27 PM
Hi
It’s not that different than a lot of components dying. They often get smoky and stinky at their
end of life. Other than the HP parts, pretty much every OCXO out there is a “no fuse” version.
Most OEM’s didn’t like the idea. The military very much did not like the idea…..
Bob
Hi
It’s not that different than a lot of components dying. They often get smoky and stinky at their
end of life. Other than the HP parts, pretty much every OCXO out there is a “no fuse” version.
Most OEM’s didn’t like the idea. The military very much did not like the idea…..
Bob
> On Mar 30, 2023, at 12:51 PM, ed breya via time-nuts <time-nuts@lists.febo.com> wrote:
>
> Wow - that is scary-burnt. Maybe it's a good idea to have some form of thermal cutout, even if a nuisance sometimes.
>
> Ed
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
AG
Adrian Godwin
Thu, Mar 30, 2023 8:07 PM
'A lot of components' don't have heaters in them, though. That's what makes
them exceptionally smoky.
On Thu, Mar 30, 2023 at 8:34 PM Bob Camp via time-nuts <
time-nuts@lists.febo.com> wrote:
Hi
It’s not that different than a lot of components dying. They often get
smoky and stinky at their
end of life. Other than the HP parts, pretty much every OCXO out there is
a “no fuse” version.
Most OEM’s didn’t like the idea. The military very much did not like the
idea…..
Bob
On Mar 30, 2023, at 12:51 PM, ed breya via time-nuts <
Wow - that is scary-burnt. Maybe it's a good idea to have some form of
thermal cutout, even if a nuisance sometimes.
'A lot of components' don't have heaters in them, though. That's what makes
them exceptionally smoky.
On Thu, Mar 30, 2023 at 8:34 PM Bob Camp via time-nuts <
time-nuts@lists.febo.com> wrote:
> Hi
>
> It’s not that different than a lot of components dying. They often get
> smoky and stinky at their
> end of life. Other than the HP parts, pretty much every OCXO out there is
> a “no fuse” version.
> Most OEM’s didn’t like the idea. The military very much did not like the
> idea…..
>
> Bob
>
> > On Mar 30, 2023, at 12:51 PM, ed breya via time-nuts <
> time-nuts@lists.febo.com> wrote:
> >
> > Wow - that is scary-burnt. Maybe it's a good idea to have some form of
> thermal cutout, even if a nuisance sometimes.
> >
> > Ed
> > _______________________________________________
> > time-nuts mailing list -- time-nuts@lists.febo.com
> > To unsubscribe send an email to time-nuts-leave@lists.febo.com
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
GW
Gary Woods
Fri, Mar 31, 2023 2:43 PM
On Thu, 30 Mar 2023 21:07:18 +0100, you wrote:
A lot of components' don't have heaters in them, though. That's what makes
them exceptionally smoky
And the silicon devices sometimes revert to sand.
Gary Woods O- K2AHC
On Thu, 30 Mar 2023 21:07:18 +0100, you wrote:
>A lot of components' don't have heaters in them, though. That's what makes
>them exceptionally smoky
And the silicon devices sometimes revert to sand.
--
Gary Woods O- K2AHC
EB
ed breya
Tue, May 9, 2023 7:33 PM
It's a bit tricky to use prescalers for arbitrary input signals. As
others have mentioned, they self-oscillate when there isn't a valid
input signal. This is not only a possible, but nearly certain
characteristic due to the (typically) ECL input amplifier being biased
right at the logic threshold, followed by lots of gain, then divided by
flip-flops. These were typically used in PLL synthesizers over limited
frequency ranges, with known, adequate power levels.
If you are counting strong signals at well defined DC levels, you can
use ECL directly with proper biasing. For a general purpose counter, you
want lots of sensitivity and wide frequency range, so the usual
prescaler with self-biasing and AC coupling is a simple way to go. You
just have to be aware of the limitations. You can eliminate the
oscillation tendency by giving up some sensitivity, but it's easiest to
just ignore it unless it causes trouble.
I'm quite familiar with the MB series and other prescalers, and have
used them in many projects. All the prescalers I have were salvaged from
old equipment, and I believe that most are long obsolete, since their
function has been mostly absorbed into the modern PLL chips.
Nowadays you can get 100-series ECL parts that can toggle to several
GHz, and in various gates, FFs, and counter types. You could build say,
a divide by 10 or other values with a programmable counter, I believe up
to a couple GHz. I don't recall all the parts and specs, but they're out
there. If you want to make it for arbitrary inputs, you'd first buffer
it with an AC coupled, self-biased line receiver cascade before
counting, and you'd be right back to a self-oscillating - but much more
versatile - prescaler.
So, if you can get this fully assembled, connectorized, ready to go
MB506 module for seven pounds, just do it - maybe get several - if the
specs are suitable for your needs. It beats the hell out a DIY one,
except for not having a nice convenient decade divide ratio. You should
look up the MB506 datasheet to see the actual capabilities - you can
usually extend the low end, for instance, by adding more input coupling
capacitance. Anyway, I don't know if the MB506 is even made anymore, but
if it is, I'd guess it alone would cost more than this module.
There used to be all sorts of prescaler ICs with various ratios,
including 10, so you might find NOS parts, but with the grief of
building something. There's also an old trick in doing counters with
binary prescalers, which is to change the counter clock frequency so the
net overall divide ratio is a nice decade value, but making the clock
and working the decimal points gets tricky. Also, if the binary divide
is big, the counter is correspondingly very slow.
Of course, if you want to get fancier, you can make a PLL/VCO based
converter that replicates the unknown input frequency at say 1/10th. But
you'll still likely have the same sort of prescaler deal involved,
inside of a PLL chip.
Ed
It's a bit tricky to use prescalers for arbitrary input signals. As
others have mentioned, they self-oscillate when there isn't a valid
input signal. This is not only a possible, but nearly certain
characteristic due to the (typically) ECL input amplifier being biased
right at the logic threshold, followed by lots of gain, then divided by
flip-flops. These were typically used in PLL synthesizers over limited
frequency ranges, with known, adequate power levels.
If you are counting strong signals at well defined DC levels, you can
use ECL directly with proper biasing. For a general purpose counter, you
want lots of sensitivity and wide frequency range, so the usual
prescaler with self-biasing and AC coupling is a simple way to go. You
just have to be aware of the limitations. You can eliminate the
oscillation tendency by giving up some sensitivity, but it's easiest to
just ignore it unless it causes trouble.
I'm quite familiar with the MB series and other prescalers, and have
used them in many projects. All the prescalers I have were salvaged from
old equipment, and I believe that most are long obsolete, since their
function has been mostly absorbed into the modern PLL chips.
Nowadays you can get 100-series ECL parts that can toggle to several
GHz, and in various gates, FFs, and counter types. You could build say,
a divide by 10 or other values with a programmable counter, I believe up
to a couple GHz. I don't recall all the parts and specs, but they're out
there. If you want to make it for arbitrary inputs, you'd first buffer
it with an AC coupled, self-biased line receiver cascade before
counting, and you'd be right back to a self-oscillating - but much more
versatile - prescaler.
So, if you can get this fully assembled, connectorized, ready to go
MB506 module for seven pounds, just do it - maybe get several - if the
specs are suitable for your needs. It beats the hell out a DIY one,
except for not having a nice convenient decade divide ratio. You should
look up the MB506 datasheet to see the actual capabilities - you can
usually extend the low end, for instance, by adding more input coupling
capacitance. Anyway, I don't know if the MB506 is even made anymore, but
if it is, I'd guess it alone would cost more than this module.
There used to be all sorts of prescaler ICs with various ratios,
including 10, so you might find NOS parts, but with the grief of
building something. There's also an old trick in doing counters with
binary prescalers, which is to change the counter clock frequency so the
net overall divide ratio is a nice decade value, but making the clock
and working the decimal points gets tricky. Also, if the binary divide
is big, the counter is correspondingly very slow.
Of course, if you want to get fancier, you can make a PLL/VCO based
converter that replicates the unknown input frequency at say 1/10th. But
you'll still likely have the same sort of prescaler deal involved,
inside of a PLL chip.
Ed
EB
ed breya
Wed, May 10, 2023 8:27 PM
Looks like the classic MC12080 is still in production by Onsemi in the
USD7 range. It can divide by 10, 20, 40, or 80 up to 1.1 GHz spec and
likely can reach over 1.5 GHz. You'd have to build a circuit for it,
unless someone offers a module of some sort.
Ed
Looks like the classic MC12080 is still in production by Onsemi in the
USD7 range. It can divide by 10, 20, 40, or 80 up to 1.1 GHz spec and
likely can reach over 1.5 GHz. You'd have to build a circuit for it,
unless someone offers a module of some sort.
Ed
EB
ed breya
Fri, May 12, 2023 5:13 PM
There are lots of newer, high performance prescalers available too, for
the much higher frequencies - especially in GaAs and other technologies.
Check this out:
https://www.analog.com/media/en/technical-documentation/data-sheets/hmc493.pdf
Note that the architecture is basically the same as in the oldies. With
the right combination of prescaler stages and baseband counting, you
could conceivably build yourself a fairly decent 18 GHz CW microwave
counter. Modern parts that are in production usually have evaluation
boards available, so putting something together isn't too hard (but
costs extra for convenient packaging).
Ed
There are lots of newer, high performance prescalers available too, for
the much higher frequencies - especially in GaAs and other technologies.
Check this out:
https://www.analog.com/media/en/technical-documentation/data-sheets/hmc493.pdf
Note that the architecture is basically the same as in the oldies. With
the right combination of prescaler stages and baseband counting, you
could conceivably build yourself a fairly decent 18 GHz CW microwave
counter. Modern parts that are in production usually have evaluation
boards available, so putting something together isn't too hard (but
costs extra for convenient packaging).
Ed
D
Dave
Fri, May 12, 2023 6:46 PM
Thanks all for the responses, given me a lot to think about.
Regards,
Dave
On 12/05/2023 18:13, ed breya via time-nuts wrote:
There are lots of newer, high performance prescalers available too, for
the much higher frequencies - especially in GaAs and other technologies.
Check this out:
https://www.analog.com/media/en/technical-documentation/data-sheets/hmc493.pdf
Note that the architecture is basically the same as in the oldies. With
the right combination of prescaler stages and baseband counting, you
could conceivably build yourself a fairly decent 18 GHz CW microwave
counter. Modern parts that are in production usually have evaluation
boards available, so putting something together isn't too hard (but
costs extra for convenient packaging).
Ed
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Thanks all for the responses, given me a lot to think about.
Regards,
Dave
On 12/05/2023 18:13, ed breya via time-nuts wrote:
> There are lots of newer, high performance prescalers available too, for
> the much higher frequencies - especially in GaAs and other technologies.
> Check this out:
>
> https://www.analog.com/media/en/technical-documentation/data-sheets/hmc493.pdf
>
> Note that the architecture is basically the same as in the oldies. With
> the right combination of prescaler stages and baseband counting, you
> could conceivably build yourself a fairly decent 18 GHz CW microwave
> counter. Modern parts that are in production usually have evaluation
> boards available, so putting something together isn't too hard (but
> costs extra for convenient packaging).
>
> Ed
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
R(
Richard (Rick) Karlquist
Fri, May 12, 2023 6:52 PM
I worked for the HP Santa Clara Division frequency counter
R&D section in the mid-1980's. Although I was not personally
involved in designing microwave counters, I did pick up
a considerable amount of knowledge about them. The idea
that a simple frequency divider could replace the
traditional harmonic sampler architecture used in microwave
counters is way too simplistic. Especially if that divider
is the dynamic variety, which is the only kind available
on the open market, AFAIK. Trying to take the entire LF to
18 GHz spectrum, and somehow process it as a single
broadband signal without filtering, and then trigger a
flip flop is not realistic, as a general purpose piece of
test equipment, when there is noise and spurs that have
to be tolerated. The HP 5340, circa 1970, actually
achieved a single coaxial input that could measure
any frequency from 10 Hz to 18 GHz, automatically.
I would venture to say that this 50+ year old technology
could still handily beat any prescaler IC today. The
project manager of the 5340 happened to sit in the
cubicle next to me.
Rick N6RK
On 5/12/2023 10:13 AM, ed breya via time-nuts wrote:
There are lots of newer, high performance prescalers available too, for
the much higher frequencies - especially in GaAs and other technologies.
Check this out:
https://www.analog.com/media/en/technical-documentation/data-sheets/hmc493.pdf
Note that the architecture is basically the same as in the oldies. With
the right combination of prescaler stages and baseband counting, you
could conceivably build yourself a fairly decent 18 GHz CW microwave
counter. Modern parts that are in production usually have evaluation
boards available, so putting something together isn't too hard (but
costs extra for convenient packaging).
Ed
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
I worked for the HP Santa Clara Division frequency counter
R&D section in the mid-1980's. Although I was not personally
involved in designing microwave counters, I did pick up
a considerable amount of knowledge about them. The idea
that a simple frequency divider could replace the
traditional harmonic sampler architecture used in microwave
counters is way too simplistic. Especially if that divider
is the dynamic variety, which is the only kind available
on the open market, AFAIK. Trying to take the entire LF to
18 GHz spectrum, and somehow process it as a single
broadband signal without filtering, and then trigger a
flip flop is not realistic, as a general purpose piece of
test equipment, when there is noise and spurs that have
to be tolerated. The HP 5340, circa 1970, actually
achieved a single coaxial input that could measure
any frequency from 10 Hz to 18 GHz, automatically.
I would venture to say that this 50+ year old technology
could still handily beat any prescaler IC today. The
project manager of the 5340 happened to sit in the
cubicle next to me.
Rick N6RK
On 5/12/2023 10:13 AM, ed breya via time-nuts wrote:
> There are lots of newer, high performance prescalers available too, for
> the much higher frequencies - especially in GaAs and other technologies.
> Check this out:
>
> https://www.analog.com/media/en/technical-documentation/data-sheets/hmc493.pdf
>
> Note that the architecture is basically the same as in the oldies. With
> the right combination of prescaler stages and baseband counting, you
> could conceivably build yourself a fairly decent 18 GHz CW microwave
> counter. Modern parts that are in production usually have evaluation
> boards available, so putting something together isn't too hard (but
> costs extra for convenient packaging).
>
> Ed
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
>
EB
ed breya
Fri, May 12, 2023 10:08 PM
Yes of course this simple kind of arrangement can't handle complicated
signals with modulation, lots of spurs or pulsed bursts and such, like a
commercial product. As I recall, the 5340 can discriminate between (pick
the largest of) spurs only a few dB apart. BTW I happen to be trying to
fix a 5340, and need some help figuring out the adjustment process. More
on that later.
A prescaler should be OK with a CW signal that's totally dominant, and
within the acceptable power range. It's just another divider that's very
fast. If you've got something like an oscillator or generator signal to
count, it should be good to go.
Ed
Yes of course this simple kind of arrangement can't handle complicated
signals with modulation, lots of spurs or pulsed bursts and such, like a
commercial product. As I recall, the 5340 can discriminate between (pick
the largest of) spurs only a few dB apart. BTW I happen to be trying to
fix a 5340, and need some help figuring out the adjustment process. More
on that later.
A prescaler should be OK with a CW signal that's totally dominant, and
within the acceptable power range. It's just another divider that's very
fast. If you've got something like an oscillator or generator signal to
count, it should be good to go.
Ed
EB
ed breya
Sat, May 13, 2023 1:00 AM
Rick, I don't see how you have the impression that any of these
prescalers are "dynamic" dividers. I remember those dynamic divider
circuits used in the old days to make 1 MHz and 100 kHz from 5 MHz, or
the more modern kind based on mixers, filters, and amplifiers, to say
make 5 from 10 Mhz, but I'm pretty sure I have never seen a
VHF/UHF/microwave prescaler IC that was not completely static - they are
chains of flip-flops and logic, pure and simple. The way they are
typically used may hide their nature a bit, but they are fundamentally
static counting machines.
Perhaps we have different definitions for dynamic vs static. My
definition of static is that there's a unique, stable machine state for
each possible count value, and you can stop it at any count value and
restart it, with the right logic activity, over any period of time. The
"right logic activity" would be to take logical control of the inputs
rather than set them up for counting RF directly in an analog
self-biased mode. In other words, you could put a gate in front of the
input to control the counting.
For instance, in the data sheet for the HMC493 it says in the General
Description:
"The HMC493LP3 & HMC493LP3E are low noise
Divide-by-4 Static Dividers utilizing InGaP GaAs HBT
technology packaged in leadless 3x3 mm QFN sur-
face mount plastic packages. This device operates
from DC (with a square wave input) to 18 GHz input
frequency from a single +5V DC supply."
The part about operating at DC "with a square wave input" means at the
normal logic levels of the device, I'm pretty sure. And usually there's
a minimum toggle edge speed for the various logic types, so there is a
little bit of dynamic-ness there, if you were to try to count at very
low rates.
Ed
Rick, I don't see how you have the impression that any of these
prescalers are "dynamic" dividers. I remember those dynamic divider
circuits used in the old days to make 1 MHz and 100 kHz from 5 MHz, or
the more modern kind based on mixers, filters, and amplifiers, to say
make 5 from 10 Mhz, but I'm pretty sure I have never seen a
VHF/UHF/microwave prescaler IC that was not completely static - they are
chains of flip-flops and logic, pure and simple. The way they are
typically used may hide their nature a bit, but they are fundamentally
static counting machines.
Perhaps we have different definitions for dynamic vs static. My
definition of static is that there's a unique, stable machine state for
each possible count value, and you can stop it at any count value and
restart it, with the right logic activity, over any period of time. The
"right logic activity" would be to take logical control of the inputs
rather than set them up for counting RF directly in an analog
self-biased mode. In other words, you could put a gate in front of the
input to control the counting.
For instance, in the data sheet for the HMC493 it says in the General
Description:
"The HMC493LP3 & HMC493LP3E are low noise
Divide-by-4 Static Dividers utilizing InGaP GaAs HBT
technology packaged in leadless 3x3 mm QFN sur-
face mount plastic packages. This device operates
from DC (with a square wave input) to 18 GHz input
frequency from a single +5V DC supply."
The part about operating at DC "with a square wave input" means at the
normal logic levels of the device, I'm pretty sure. And usually there's
a minimum toggle edge speed for the various logic types, so there is a
little bit of dynamic-ness there, if you were to try to count at very
low rates.
Ed
R(
Richard (Rick) Karlquist
Sat, May 13, 2023 4:02 PM
On 5/12/2023 6:00 PM, ed breya via time-nuts wrote:
Rick, I don't see how you have the impression that any of these
prescalers are "dynamic" dividers. \
That is what the project manager of the HP 5386 told me back in
the 80's. He used a static divider made by HP Santa Rosa. The
5334A used a dynamic divider made by HP Santa Clara. As the
5334B project manager, I tried whatever COTS dividers I could
find and they all had the same problems as the divider made
in Santa Clara. I designed that one out to save $100.
You previously said something like any prescaler will work if
the signal is coming from a signal generator. This notion is
contradicted by the HP 8660. For both the 5334A and 5334B,
the test procedure for the C channel is to drive the counter
with an 8660 at 1.3 GHz to verify the specified bandwidth.
However, the TP goes on to say that a 1 GHz high pass filter
must be inserted in front of the counter. And sure enough,
if you don't do this, it won't count 1.3 GHz. Admittedly,
the 8660 was not exactly what you would call HP's "finest
hour." I suspect you could dispense with the filter if you
used an 8662/3.
I was not aware of the HMC493, which seems to be static. With
that chip you might finally be able to homebrew a decent C channel.
The description of the technology used sounds like something
that would come out of the Santa Rosa IC fab. The price is a
bargain IMHO at $35. For $425 you can get an eval board.
So it looks like the technology has advanced somewhat in
35 years.
Rick
I remember those dynamic divider
circuits used in the old days to make 1 MHz and 100 kHz from 5 MHz, or
the more modern kind based on mixers, filters, and amplifiers, to say
make 5 from 10 Mhz, but I'm pretty sure I have never seen a
VHF/UHF/microwave prescaler IC that was not completely static - they are
chains of flip-flops and logic, pure and simple. The way they are
typically used may hide their nature a bit, but they are fundamentally
static counting machines.
Perhaps we have different definitions for dynamic vs static. My
definition of static is that there's a unique, stable machine state for
each possible count value, and you can stop it at any count value and
restart it, with the right logic activity, over any period of time. The
"right logic activity" would be to take logical control of the inputs
rather than set them up for counting RF directly in an analog
self-biased mode. In other words, you could put a gate in front of the
input to control the counting.
For instance, in the data sheet for the HMC493 it says in the General
Description:
"The HMC493LP3 & HMC493LP3E are low noise
Divide-by-4 Static Dividers utilizing InGaP GaAs HBT
technology packaged in leadless 3x3 mm QFN sur-
face mount plastic packages. This device operates
from DC (with a square wave input) to 18 GHz input
frequency from a single +5V DC supply."
The part about operating at DC "with a square wave input" means at the
normal logic levels of the device, I'm pretty sure. And usually there's
a minimum toggle edge speed for the various logic types, so there is a
little bit of dynamic-ness there, if you were to try to count at very
low rates.
Ed
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
On 5/12/2023 6:00 PM, ed breya via time-nuts wrote:
> Rick, I don't see how you have the impression that any of these
> prescalers are "dynamic" dividers. \
That is what the project manager of the HP 5386 told me back in
the 80's. He used a static divider made by HP Santa Rosa. The
5334A used a dynamic divider made by HP Santa Clara. As the
5334B project manager, I tried whatever COTS dividers I could
find and they all had the same problems as the divider made
in Santa Clara. I designed that one out to save $100.
You previously said something like any prescaler will work if
the signal is coming from a signal generator. This notion is
contradicted by the HP 8660. For both the 5334A and 5334B,
the test procedure for the C channel is to drive the counter
with an 8660 at 1.3 GHz to verify the specified bandwidth.
However, the TP goes on to say that a 1 GHz high pass filter
must be inserted in front of the counter. And sure enough,
if you don't do this, it won't count 1.3 GHz. Admittedly,
the 8660 was not exactly what you would call HP's "finest
hour." I suspect you could dispense with the filter if you
used an 8662/3.
I was not aware of the HMC493, which seems to be static. With
that chip you might finally be able to homebrew a decent C channel.
The description of the technology used sounds like something
that would come out of the Santa Rosa IC fab. The price is a
bargain IMHO at $35. For $425 you can get an eval board.
So it looks like the technology has advanced somewhat in
35 years.
Rick
I remember those dynamic divider
> circuits used in the old days to make 1 MHz and 100 kHz from 5 MHz, or
> the more modern kind based on mixers, filters, and amplifiers, to say
> make 5 from 10 Mhz, but I'm pretty sure I have never seen a
> VHF/UHF/microwave prescaler IC that was not completely static - they are
> chains of flip-flops and logic, pure and simple. The way they are
> typically used may hide their nature a bit, but they are fundamentally
> static counting machines.
>
> Perhaps we have different definitions for dynamic vs static. My
> definition of static is that there's a unique, stable machine state for
> each possible count value, and you can stop it at any count value and
> restart it, with the right logic activity, over any period of time. The
> "right logic activity" would be to take logical control of the inputs
> rather than set them up for counting RF directly in an analog
> self-biased mode. In other words, you could put a gate in front of the
> input to control the counting.
>
> For instance, in the data sheet for the HMC493 it says in the General
> Description:
>
> "The HMC493LP3 & HMC493LP3E are low noise
> Divide-by-4 Static Dividers utilizing InGaP GaAs HBT
> technology packaged in leadless 3x3 mm QFN sur-
> face mount plastic packages. This device operates
> from DC (with a square wave input) to 18 GHz input
> frequency from a single +5V DC supply."
>
> The part about operating at DC "with a square wave input" means at the
> normal logic levels of the device, I'm pretty sure. And usually there's
> a minimum toggle edge speed for the various logic types, so there is a
> little bit of dynamic-ness there, if you were to try to count at very
> low rates.
>
> Ed
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
>
G
ghf (at) hoffmann-hochfrequenz.de
Sat, May 13, 2023 5:21 PM
Am 2023-05-13 3:00, schrieb ed breya via time-nuts:
"The HMC493LP3 & HMC493LP3E are low noise
Divide-by-4 Static Dividers utilizing InGaP GaAs HBT
I have built a 1/8 prescaler using the HMC363 to extend the trigger
bandwidth of my 54750A scope from 2.5 to 12 GHz (or the SR620).
It was necessary to have a gain block in front of the prescaler
since the HMC363 pumped some of the divided signal back into its input
which contaminated the measured channel with some jitter.
A second gain block at the output makes sure that there is enough
signal left even after a power divider.
The home-etched board is still without the preamplifier.
Oscillation tendency is most easily cured by a small DC offset.
That costs some sensitivity, another reason for the preamp.
The design has been published in DUBUS last year. I'll put it
on my web site when the FTP access works again. Gerbers are available.
Cheers, Gerhard DK4XP
Am 2023-05-13 3:00, schrieb ed breya via time-nuts:
> "The HMC493LP3 & HMC493LP3E are low noise
> Divide-by-4 Static Dividers utilizing InGaP GaAs HBT
I have built a 1/8 prescaler using the HMC363 to extend the trigger
bandwidth of my 54750A scope from 2.5 to 12 GHz (or the SR620).
It was necessary to have a gain block in front of the prescaler
since the HMC363 pumped some of the divided signal back into its input
which contaminated the measured channel with some jitter.
A second gain block at the output makes sure that there is enough
signal left even after a power divider.
The home-etched board is still without the preamplifier.
Oscillation tendency is most easily cured by a small DC offset.
That costs some sensitivity, another reason for the preamp.
The design has been published in DUBUS last year. I'll put it
on my web site when the FTP access works again. Gerbers are available.
Cheers, Gerhard DK4XP
LJ
Lux, Jim
Sat, May 13, 2023 5:55 PM
On 5/13/23 9:02 AM, Richard (Rick) Karlquist via time-nuts wrote:
On 5/12/2023 6:00 PM, ed breya via time-nuts wrote:
Rick, I don't see how you have the impression that any of these
prescalers are "dynamic" dividers. \
That is what the project manager of the HP 5386 told me back in
the 80's. He used a static divider made by HP Santa Rosa. The
5334A used a dynamic divider made by HP Santa Clara. As the
5334B project manager, I tried whatever COTS dividers I could
find and they all had the same problems as the divider made
in Santa Clara. I designed that one out to save $100.
You previously said something like any prescaler will work if
the signal is coming from a signal generator. This notion is
contradicted by the HP 8660. For both the 5334A and 5334B,
the test procedure for the C channel is to drive the counter
with an 8660 at 1.3 GHz to verify the specified bandwidth.
However, the TP goes on to say that a 1 GHz high pass filter
must be inserted in front of the counter. And sure enough,
if you don't do this, it won't count 1.3 GHz. Admittedly,
the 8660 was not exactly what you would call HP's "finest
hour." I suspect you could dispense with the filter if you
used an 8662/3.
I was not aware of the HMC493, which seems to be static. With
that chip you might finally be able to homebrew a decent C channel.
The description of the technology used sounds like something
that would come out of the Santa Rosa IC fab. The price is a
bargain IMHO at $35. For $425 you can get an eval board.
So it looks like the technology has advanced somewhat in
35 years.
Hittite (HMC) made a bunch of GaAs dividers (and some are still being
made) that worked up at microwave frequencies. Some of them were
originally developed under SBIR funding from NASA. We use them in
synthesizers for deep space radios. We were trying to get wideband
(50-100 MHz) tuning coverage at X-band, and our ideas with DROs with two
varactors didn't work out. So we went to MMIC VCO and dividers in a
conventional PLL sort of arrangement (as opposed to DROs and SPDs)
https://ipnpr.jpl.nasa.gov/progress_report/42-166/166A.pdf
describes the approach and measured performance in the application
(using eval boards). I can't remember what parts we used for the divide
by 8 and the divide by N, and the picture is too low res to read it off
the part. I can probably find it. (Oddly, I think I saw these
breadboards a couple years ago in a closet).
The Iris cubesat transponder uses a divide by 2 like the HMC361 from 8.5
GHz or 7.1 GHz
https://digitalcommons.usu.edu/cgi/viewcontent.cgi?article=3603&context=smallsat
It's true that in these applications, the divider is seeing only one
signal with very good SNR - it's not like a general purpose counter.
As far as I know, they're just a chain of flipflops, sometimes with some
logic (there was a variable divide ratio one we used). They aren't
exactly low power.
On 5/13/23 9:02 AM, Richard (Rick) Karlquist via time-nuts wrote:
>
>
> On 5/12/2023 6:00 PM, ed breya via time-nuts wrote:
>> Rick, I don't see how you have the impression that any of these
>> prescalers are "dynamic" dividers. \
>
> That is what the project manager of the HP 5386 told me back in
> the 80's. He used a static divider made by HP Santa Rosa. The
> 5334A used a dynamic divider made by HP Santa Clara. As the
> 5334B project manager, I tried whatever COTS dividers I could
> find and they all had the same problems as the divider made
> in Santa Clara. I designed that one out to save $100.
>
> You previously said something like any prescaler will work if
> the signal is coming from a signal generator. This notion is
> contradicted by the HP 8660. For both the 5334A and 5334B,
> the test procedure for the C channel is to drive the counter
> with an 8660 at 1.3 GHz to verify the specified bandwidth.
> However, the TP goes on to say that a 1 GHz high pass filter
> must be inserted in front of the counter. And sure enough,
> if you don't do this, it won't count 1.3 GHz. Admittedly,
> the 8660 was not exactly what you would call HP's "finest
> hour." I suspect you could dispense with the filter if you
> used an 8662/3.
>
> I was not aware of the HMC493, which seems to be static. With
> that chip you might finally be able to homebrew a decent C channel.
> The description of the technology used sounds like something
> that would come out of the Santa Rosa IC fab. The price is a
> bargain IMHO at $35. For $425 you can get an eval board.
> So it looks like the technology has advanced somewhat in
> 35 years.
Hittite (HMC) made a bunch of GaAs dividers (and some are still being
made) that worked up at microwave frequencies. Some of them were
originally developed under SBIR funding from NASA. We use them in
synthesizers for deep space radios. We were trying to get wideband
(50-100 MHz) tuning coverage at X-band, and our ideas with DROs with two
varactors didn't work out. So we went to MMIC VCO and dividers in a
conventional PLL sort of arrangement (as opposed to DROs and SPDs)
https://ipnpr.jpl.nasa.gov/progress_report/42-166/166A.pdf
describes the approach and measured performance in the application
(using eval boards). I can't remember what parts we used for the divide
by 8 and the divide by N, and the picture is too low res to read it off
the part. I can probably find it. (Oddly, I think I saw these
breadboards a couple years ago in a closet).
The Iris cubesat transponder uses a divide by 2 like the HMC361 from 8.5
GHz or 7.1 GHz
https://digitalcommons.usu.edu/cgi/viewcontent.cgi?article=3603&context=smallsat
It's true that in these applications, the divider is seeing only one
signal with very good SNR - it's not like a general purpose counter.
As far as I know, they're just a chain of flipflops, sometimes with some
logic (there was a variable divide ratio one we used). They aren't
exactly low power.
EB
ed breya
Sat, May 13, 2023 9:16 PM
I think the very first prescaler IC I ever encountered, long ago, is the
11C90. I believe it's heyday was in the 1980s - it appears it was near
EOL by 1992. I discovered it in some junk 6 GHz synthesizer modules that
used a PLLed VHF VCO and a power amplifier, that was then SRD/BPF
multiplied up the rest of the way. Later, I found one used as a
prescaler in a low cost frequency counter from the 80s or 90s.
A data sheet is here - it provides some more perspective on how these
work, and their evolution.
https://pdf.datasheetcatalog.com/datasheet/nationalsemiconductor/11C90.pdf
You can see that it is basically a high speed ECL 1/10 divider (or 1/11
- dual modulus even) capable of DC to around 650 MHz. It also includes
provision to connect it for self-biased clock input at the logic
threshold for AC coupling without any other parts (except an input
capacitor depending on lowest frequency needed). This made it possible
to toggle with input levels a fair amount less than the normal ECL
range. I think the later, improved technologies used more gain stages to
get the higher sensitivities we see in the modern (1990s-present) types.
There is of course a trade-off between speed and sensitivity, since more
gain slows it down.
Anyway, if you thoroughly read the Functional Description on page 6,
you'll see the story from the horse's mouth, so to speak. BTW I believe
there is a typo in the second paragraph, where it says
"This feature ensures that the circuit will operate with clock waveforms
having very slow rise and fall times, and thus, there is no maximum
frequency restriction."
I think it was meant to say no "minimum" frequency restriction, which
makes more sense, in context.
I think around here is where prescalers were becoming a distinct
functional component class, more tailored to the huge emerging
telecommunications markets - optimized for RF, and separate from the
counters category, although used together. At heart, they are the same.
Enjoy the story.
Ed
I think the very first prescaler IC I ever encountered, long ago, is the
11C90. I believe it's heyday was in the 1980s - it appears it was near
EOL by 1992. I discovered it in some junk 6 GHz synthesizer modules that
used a PLLed VHF VCO and a power amplifier, that was then SRD/BPF
multiplied up the rest of the way. Later, I found one used as a
prescaler in a low cost frequency counter from the 80s or 90s.
A data sheet is here - it provides some more perspective on how these
work, and their evolution.
https://pdf.datasheetcatalog.com/datasheet/nationalsemiconductor/11C90.pdf
You can see that it is basically a high speed ECL 1/10 divider (or 1/11
- dual modulus even) capable of DC to around 650 MHz. It also includes
provision to connect it for self-biased clock input at the logic
threshold for AC coupling without any other parts (except an input
capacitor depending on lowest frequency needed). This made it possible
to toggle with input levels a fair amount less than the normal ECL
range. I think the later, improved technologies used more gain stages to
get the higher sensitivities we see in the modern (1990s-present) types.
There is of course a trade-off between speed and sensitivity, since more
gain slows it down.
Anyway, if you thoroughly read the Functional Description on page 6,
you'll see the story from the horse's mouth, so to speak. BTW I believe
there is a typo in the second paragraph, where it says
"This feature ensures that the circuit will operate with clock waveforms
having very slow rise and fall times, and thus, there is no maximum
frequency restriction."
I think it was meant to say no "minimum" frequency restriction, which
makes more sense, in context.
I think around here is where prescalers were becoming a distinct
functional component class, more tailored to the huge emerging
telecommunications markets - optimized for RF, and separate from the
counters category, although used together. At heart, they are the same.
Enjoy the story.
Ed
BC
Bob Camp
Sat, May 13, 2023 11:58 PM
Hi
The “junk back on the input” is not at all uncommon. There are a number of PLL
IC’s that have the same “feature”. Buffering the input signal is often a good idea.
Bob
On May 13, 2023, at 1:21 PM, Gerhard Hoffmann via time-nuts time-nuts@lists.febo.com wrote:
Am 2023-05-13 3:00, schrieb ed breya via time-nuts:
"The HMC493LP3 & HMC493LP3E are low noise
Divide-by-4 Static Dividers utilizing InGaP GaAs HBT
I have built a 1/8 prescaler using the HMC363 to extend the trigger
bandwidth of my 54750A scope from 2.5 to 12 GHz (or the SR620).
It was necessary to have a gain block in front of the prescaler
since the HMC363 pumped some of the divided signal back into its input
which contaminated the measured channel with some jitter.
A second gain block at the output makes sure that there is enough
signal left even after a power divider.
The home-etched board is still without the preamplifier.
Oscillation tendency is most easily cured by a small DC offset.
That costs some sensitivity, another reason for the preamp.
The design has been published in DUBUS last year. I'll put it
on my web site when the FTP access works again. Gerbers are available.
Cheers, Gerhard DK4XP<Auswahl_2023051318:57:49.png><Auswahl_2023051319:12:15.png>_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Hi
The “junk back on the input” is not at all uncommon. There are a number of PLL
IC’s that have the same “feature”. Buffering the input signal is often a good idea.
Bob
> On May 13, 2023, at 1:21 PM, Gerhard Hoffmann via time-nuts <time-nuts@lists.febo.com> wrote:
>
> Am 2023-05-13 3:00, schrieb ed breya via time-nuts:
>
>> "The HMC493LP3 & HMC493LP3E are low noise
>> Divide-by-4 Static Dividers utilizing InGaP GaAs HBT
>
> I have built a 1/8 prescaler using the HMC363 to extend the trigger
> bandwidth of my 54750A scope from 2.5 to 12 GHz (or the SR620).
>
> It was necessary to have a gain block in front of the prescaler
> since the HMC363 pumped some of the divided signal back into its input
> which contaminated the measured channel with some jitter.
> A second gain block at the output makes sure that there is enough
> signal left even after a power divider.
>
> The home-etched board is still without the preamplifier.
> Oscillation tendency is most easily cured by a small DC offset.
> That costs some sensitivity, another reason for the preamp.
>
> The design has been published in DUBUS last year. I'll put it
> on my web site when the FTP access works again. Gerbers are available.
>
> Cheers, Gerhard DK4XP<Auswahl_2023051318:57:49.png><Auswahl_2023051319:12:15.png>_______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com