HP Z3801A Internal Power Supply issues

AM
Artek Manuals
Sun, Jan 31, 2016 2:21 AM

My Z3801A began to die over the last few days. It would work for an hour
or two then produce a receiver error and communications problems. Turn
it off, let it sit awhile, back on and it would start the survey and
then begin to recover. Tonight it finally went dead with high current on
the main input . The feed supply  has a 2amp current limit so I have not
blown the two 3amp fuses

I have a schematic of the supply reverse engineered by Stefan Hegnauere
a few years ago (thank you very much) and a limited theory of operation
description so I am not totally dead from a troubleshooting point of
view although if one of the DC-DC converters is dead then I will likely
have to scrap the supply

Questions

  1. Any prior experience as where to look first I checked the
    electrolytics and at first glance they seem fine, no loss of magic smoke
    or any parts obviously hot or bulging. ESR 's look good on those
  2. Will this supply function unloaded while I troubleshoot it on the
    bench out of the unit?
  3. Anyone tried independent powering the Z3801? looks like two 5V
    supplies and +/- 15V..not hard to to do
  4. Anyone with a good supply from a parts unit want to part with same?

Dave
NR1DX
ArtekManuals.com

--
Dave
Manuals@ArtekManuals.com
www.ArtekManuals.com


This email has been checked for viruses by Avast antivirus software.
https://www.avast.com/antivirus

My Z3801A began to die over the last few days. It would work for an hour or two then produce a receiver error and communications problems. Turn it off, let it sit awhile, back on and it would start the survey and then begin to recover. Tonight it finally went dead with high current on the main input . The feed supply has a 2amp current limit so I have not blown the two 3amp fuses I have a schematic of the supply reverse engineered by Stefan Hegnauere a few years ago (thank you very much) and a limited theory of operation description so I am not totally dead from a troubleshooting point of view although if one of the DC-DC converters is dead then I will likely have to scrap the supply Questions 1) Any prior experience as where to look first I checked the electrolytics and at first glance they seem fine, no loss of magic smoke or any parts obviously hot or bulging. ESR 's look good on those 2) Will this supply function unloaded while I troubleshoot it on the bench out of the unit? 3) Anyone tried independent powering the Z3801? looks like two 5V supplies and +/- 15V..not hard to to do 4) Anyone with a good supply from a parts unit want to part with same? Dave NR1DX ArtekManuals.com -- Dave Manuals@ArtekManuals.com www.ArtekManuals.com --- This email has been checked for viruses by Avast antivirus software. https://www.avast.com/antivirus
AP
Alex Pummer
Sun, Jan 31, 2016 5:04 AM

power supply runs for a while, output current increases, but voltage
stays correct -> problem is in the load circuit, load circuit working
for a while load current increases, but it recovers after pause -> look
for some tantalum capacitor in the load circuit, they recover often  a
few times before they finally die,
KJ6UHN
Alex

On 1/30/2016 6:21 PM, Artek Manuals wrote:

My Z3801A began to die over the last few days. It would work for an
hour or two then produce a receiver error and communications problems.
Turn it off, let it sit awhile, back on and it would start the survey
and then begin to recover. Tonight it finally went dead with high
current on the main input . The feed supply has a 2amp current limit
so I have not blown the two 3amp fuses

I have a schematic of the supply reverse engineered by Stefan
Hegnauere a few years ago (thank you very much) and a limited theory
of operation description so I am not totally dead from a
troubleshooting point of view although if one of the DC-DC converters
is dead then I will likely have to scrap the supply

Questions

  1. Any prior experience as where to look first I checked the
    electrolytics and at first glance they seem fine, no loss of magic
    smoke or any parts obviously hot or bulging. ESR 's look good on those
  2. Will this supply function unloaded while I troubleshoot it on the
    bench out of the unit?
  3. Anyone tried independent powering the Z3801? looks like two 5V
    supplies and +/- 15V..not hard to to do
  4. Anyone with a good supply from a parts unit want to part with same?

Dave
NR1DX
ArtekManuals.com

power supply runs for a while, output current increases, but voltage stays correct -> problem is in the load circuit, load circuit working for a while load current increases, but it recovers after pause -> look for some tantalum capacitor in the load circuit, they recover often a few times before they finally die, KJ6UHN Alex On 1/30/2016 6:21 PM, Artek Manuals wrote: > My Z3801A began to die over the last few days. It would work for an > hour or two then produce a receiver error and communications problems. > Turn it off, let it sit awhile, back on and it would start the survey > and then begin to recover. Tonight it finally went dead with high > current on the main input . The feed supply has a 2amp current limit > so I have not blown the two 3amp fuses > > I have a schematic of the supply reverse engineered by Stefan > Hegnauere a few years ago (thank you very much) and a limited theory > of operation description so I am not totally dead from a > troubleshooting point of view although if one of the DC-DC converters > is dead then I will likely have to scrap the supply > > Questions > 1) Any prior experience as where to look first I checked the > electrolytics and at first glance they seem fine, no loss of magic > smoke or any parts obviously hot or bulging. ESR 's look good on those > 2) Will this supply function unloaded while I troubleshoot it on the > bench out of the unit? > 3) Anyone tried independent powering the Z3801? looks like two 5V > supplies and +/- 15V..not hard to to do > 4) Anyone with a good supply from a parts unit want to part with same? > > Dave > NR1DX > ArtekManuals.com >
PS
paul swed
Sun, Jan 31, 2016 2:11 PM

As I mentioned in the another thread this week.
You can indeed replace the failed voltage with just another inverter.
Mine has run this way for 10 years. Mine was a 5 by the way. If it was one
side of the +/-15 I would say you want to replace both.
The entire switcher can be replaced with supplies that work off of 110 so
you can get rid of the 48 volt supplies.
Good luck
Paul
WB8TSL

On Sun, Jan 31, 2016 at 12:04 AM, Alex Pummer alex@pcscons.com wrote:

power supply runs for a while, output current increases, but voltage stays
correct -> problem is in the load circuit, load circuit working for a while
load current increases, but it recovers after pause -> look for some
tantalum capacitor in the load circuit, they recover often  a few times
before they finally die,
KJ6UHN
Alex

On 1/30/2016 6:21 PM, Artek Manuals wrote:

My Z3801A began to die over the last few days. It would work for an hour
or two then produce a receiver error and communications problems. Turn it
off, let it sit awhile, back on and it would start the survey and then
begin to recover. Tonight it finally went dead with high current on the
main input . The feed supply has a 2amp current limit so I have not blown
the two 3amp fuses

I have a schematic of the supply reverse engineered by Stefan Hegnauere a
few years ago (thank you very much) and a limited theory of operation
description so I am not totally dead from a troubleshooting point of view
although if one of the DC-DC converters is dead then I will likely have to
scrap the supply

Questions

  1. Any prior experience as where to look first I checked the
    electrolytics and at first glance they seem fine, no loss of magic smoke or
    any parts obviously hot or bulging. ESR 's look good on those
  2. Will this supply function unloaded while I troubleshoot it on the
    bench out of the unit?
  3. Anyone tried independent powering the Z3801? looks like two 5V
    supplies and +/- 15V..not hard to to do
  4. Anyone with a good supply from a parts unit want to part with same?

Dave
NR1DX
ArtekManuals.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

As I mentioned in the another thread this week. You can indeed replace the failed voltage with just another inverter. Mine has run this way for 10 years. Mine was a 5 by the way. If it was one side of the +/-15 I would say you want to replace both. The entire switcher can be replaced with supplies that work off of 110 so you can get rid of the 48 volt supplies. Good luck Paul WB8TSL On Sun, Jan 31, 2016 at 12:04 AM, Alex Pummer <alex@pcscons.com> wrote: > power supply runs for a while, output current increases, but voltage stays > correct -> problem is in the load circuit, load circuit working for a while > load current increases, but it recovers after pause -> look for some > tantalum capacitor in the load circuit, they recover often a few times > before they finally die, > KJ6UHN > Alex > > On 1/30/2016 6:21 PM, Artek Manuals wrote: > >> My Z3801A began to die over the last few days. It would work for an hour >> or two then produce a receiver error and communications problems. Turn it >> off, let it sit awhile, back on and it would start the survey and then >> begin to recover. Tonight it finally went dead with high current on the >> main input . The feed supply has a 2amp current limit so I have not blown >> the two 3amp fuses >> >> I have a schematic of the supply reverse engineered by Stefan Hegnauere a >> few years ago (thank you very much) and a limited theory of operation >> description so I am not totally dead from a troubleshooting point of view >> although if one of the DC-DC converters is dead then I will likely have to >> scrap the supply >> >> Questions >> 1) Any prior experience as where to look first I checked the >> electrolytics and at first glance they seem fine, no loss of magic smoke or >> any parts obviously hot or bulging. ESR 's look good on those >> 2) Will this supply function unloaded while I troubleshoot it on the >> bench out of the unit? >> 3) Anyone tried independent powering the Z3801? looks like two 5V >> supplies and +/- 15V..not hard to to do >> 4) Anyone with a good supply from a parts unit want to part with same? >> >> Dave >> NR1DX >> ArtekManuals.com >> >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
BC
Bob Camp
Sun, Jan 31, 2016 4:42 PM

Hi

I had one that took out the internal fuse once every 3 to 5 years. After
replacing the fuse 3 times, the supply died. Go figure … only 15 years or use.

Bob

On Jan 31, 2016, at 9:11 AM, paul swed paulswedb@gmail.com wrote:

As I mentioned in the another thread this week.
You can indeed replace the failed voltage with just another inverter.
Mine has run this way for 10 years. Mine was a 5 by the way. If it was one
side of the +/-15 I would say you want to replace both.
The entire switcher can be replaced with supplies that work off of 110 so
you can get rid of the 48 volt supplies.
Good luck
Paul
WB8TSL

On Sun, Jan 31, 2016 at 12:04 AM, Alex Pummer alex@pcscons.com wrote:

power supply runs for a while, output current increases, but voltage stays
correct -> problem is in the load circuit, load circuit working for a while
load current increases, but it recovers after pause -> look for some
tantalum capacitor in the load circuit, they recover often  a few times
before they finally die,
KJ6UHN
Alex

On 1/30/2016 6:21 PM, Artek Manuals wrote:

My Z3801A began to die over the last few days. It would work for an hour
or two then produce a receiver error and communications problems. Turn it
off, let it sit awhile, back on and it would start the survey and then
begin to recover. Tonight it finally went dead with high current on the
main input . The feed supply has a 2amp current limit so I have not blown
the two 3amp fuses

I have a schematic of the supply reverse engineered by Stefan Hegnauere a
few years ago (thank you very much) and a limited theory of operation
description so I am not totally dead from a troubleshooting point of view
although if one of the DC-DC converters is dead then I will likely have to
scrap the supply

Questions

  1. Any prior experience as where to look first I checked the
    electrolytics and at first glance they seem fine, no loss of magic smoke or
    any parts obviously hot or bulging. ESR 's look good on those
  2. Will this supply function unloaded while I troubleshoot it on the
    bench out of the unit?
  3. Anyone tried independent powering the Z3801? looks like two 5V
    supplies and +/- 15V..not hard to to do
  4. Anyone with a good supply from a parts unit want to part with same?

Dave
NR1DX
ArtekManuals.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi I had one that took out the internal fuse once every 3 to 5 years. After replacing the fuse 3 times, the supply died. Go figure … only 15 years or use. Bob > On Jan 31, 2016, at 9:11 AM, paul swed <paulswedb@gmail.com> wrote: > > As I mentioned in the another thread this week. > You can indeed replace the failed voltage with just another inverter. > Mine has run this way for 10 years. Mine was a 5 by the way. If it was one > side of the +/-15 I would say you want to replace both. > The entire switcher can be replaced with supplies that work off of 110 so > you can get rid of the 48 volt supplies. > Good luck > Paul > WB8TSL > > On Sun, Jan 31, 2016 at 12:04 AM, Alex Pummer <alex@pcscons.com> wrote: > >> power supply runs for a while, output current increases, but voltage stays >> correct -> problem is in the load circuit, load circuit working for a while >> load current increases, but it recovers after pause -> look for some >> tantalum capacitor in the load circuit, they recover often a few times >> before they finally die, >> KJ6UHN >> Alex >> >> On 1/30/2016 6:21 PM, Artek Manuals wrote: >> >>> My Z3801A began to die over the last few days. It would work for an hour >>> or two then produce a receiver error and communications problems. Turn it >>> off, let it sit awhile, back on and it would start the survey and then >>> begin to recover. Tonight it finally went dead with high current on the >>> main input . The feed supply has a 2amp current limit so I have not blown >>> the two 3amp fuses >>> >>> I have a schematic of the supply reverse engineered by Stefan Hegnauere a >>> few years ago (thank you very much) and a limited theory of operation >>> description so I am not totally dead from a troubleshooting point of view >>> although if one of the DC-DC converters is dead then I will likely have to >>> scrap the supply >>> >>> Questions >>> 1) Any prior experience as where to look first I checked the >>> electrolytics and at first glance they seem fine, no loss of magic smoke or >>> any parts obviously hot or bulging. ESR 's look good on those >>> 2) Will this supply function unloaded while I troubleshoot it on the >>> bench out of the unit? >>> 3) Anyone tried independent powering the Z3801? looks like two 5V >>> supplies and +/- 15V..not hard to to do >>> 4) Anyone with a good supply from a parts unit want to part with same? >>> >>> Dave >>> NR1DX >>> ArtekManuals.com >>> >>> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
AG
Adrian Godwin
Sun, Jan 31, 2016 9:43 PM

A tant melted on one of my KS-246361 boxes (I guess they're a similar age,
possibly even the same source). The only one so far to have trouble. It
failed in such a way that it continued to smoke and load up the power
supply,

The current drawn from the no-name laptop power supply was still well
within spec, at least by the time I post-mortem'ed it. But that melted too,
and took out three levels of circuit breakers. The KS-24361 worked
perfectly after replacing the cap.

On Sun, Jan 31, 2016 at 4:42 PM, Bob Camp kb8tq@n1k.org wrote:

Hi

I had one that took out the internal fuse once every 3 to 5 years. After
replacing the fuse 3 times, the supply died. Go figure … only 15 years or
use.

Bob

On Jan 31, 2016, at 9:11 AM, paul swed paulswedb@gmail.com wrote:

As I mentioned in the another thread this week.
You can indeed replace the failed voltage with just another inverter.
Mine has run this way for 10 years. Mine was a 5 by the way. If it was

one

side of the +/-15 I would say you want to replace both.
The entire switcher can be replaced with supplies that work off of 110 so
you can get rid of the 48 volt supplies.
Good luck
Paul
WB8TSL

On Sun, Jan 31, 2016 at 12:04 AM, Alex Pummer alex@pcscons.com wrote:

power supply runs for a while, output current increases, but voltage

stays

correct -> problem is in the load circuit, load circuit working for a

while

load current increases, but it recovers after pause -> look for some
tantalum capacitor in the load circuit, they recover often  a few times
before they finally die,
KJ6UHN
Alex

On 1/30/2016 6:21 PM, Artek Manuals wrote:

My Z3801A began to die over the last few days. It would work for an

hour

or two then produce a receiver error and communications problems. Turn

it

off, let it sit awhile, back on and it would start the survey and then
begin to recover. Tonight it finally went dead with high current on the
main input . The feed supply has a 2amp current limit so I have not

blown

the two 3amp fuses

I have a schematic of the supply reverse engineered by Stefan

Hegnauere a

few years ago (thank you very much) and a limited theory of operation
description so I am not totally dead from a troubleshooting point of

view

although if one of the DC-DC converters is dead then I will likely

have to

scrap the supply

Questions

  1. Any prior experience as where to look first I checked the
    electrolytics and at first glance they seem fine, no loss of magic

smoke or

any parts obviously hot or bulging. ESR 's look good on those
2) Will this supply function unloaded while I troubleshoot it on the
bench out of the unit?
3) Anyone tried independent powering the Z3801? looks like two 5V
supplies and +/- 15V..not hard to to do
4) Anyone with a good supply from a parts unit want to part with same?

Dave
NR1DX
ArtekManuals.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

A tant melted on one of my KS-246361 boxes (I guess they're a similar age, possibly even the same source). The only one so far to have trouble. It failed in such a way that it continued to smoke and load up the power supply, The current drawn from the no-name laptop power supply was still well within spec, at least by the time I post-mortem'ed it. But that melted too, and took out three levels of circuit breakers. The KS-24361 worked perfectly after replacing the cap. On Sun, Jan 31, 2016 at 4:42 PM, Bob Camp <kb8tq@n1k.org> wrote: > Hi > > I had one that took out the internal fuse once every 3 to 5 years. After > replacing the fuse 3 times, the supply died. Go figure … only 15 years or > use. > > Bob > > On Jan 31, 2016, at 9:11 AM, paul swed <paulswedb@gmail.com> wrote: > > > > As I mentioned in the another thread this week. > > You can indeed replace the failed voltage with just another inverter. > > Mine has run this way for 10 years. Mine was a 5 by the way. If it was > one > > side of the +/-15 I would say you want to replace both. > > The entire switcher can be replaced with supplies that work off of 110 so > > you can get rid of the 48 volt supplies. > > Good luck > > Paul > > WB8TSL > > > > On Sun, Jan 31, 2016 at 12:04 AM, Alex Pummer <alex@pcscons.com> wrote: > > > >> power supply runs for a while, output current increases, but voltage > stays > >> correct -> problem is in the load circuit, load circuit working for a > while > >> load current increases, but it recovers after pause -> look for some > >> tantalum capacitor in the load circuit, they recover often a few times > >> before they finally die, > >> KJ6UHN > >> Alex > >> > >> On 1/30/2016 6:21 PM, Artek Manuals wrote: > >> > >>> My Z3801A began to die over the last few days. It would work for an > hour > >>> or two then produce a receiver error and communications problems. Turn > it > >>> off, let it sit awhile, back on and it would start the survey and then > >>> begin to recover. Tonight it finally went dead with high current on the > >>> main input . The feed supply has a 2amp current limit so I have not > blown > >>> the two 3amp fuses > >>> > >>> I have a schematic of the supply reverse engineered by Stefan > Hegnauere a > >>> few years ago (thank you very much) and a limited theory of operation > >>> description so I am not totally dead from a troubleshooting point of > view > >>> although if one of the DC-DC converters is dead then I will likely > have to > >>> scrap the supply > >>> > >>> Questions > >>> 1) Any prior experience as where to look first I checked the > >>> electrolytics and at first glance they seem fine, no loss of magic > smoke or > >>> any parts obviously hot or bulging. ESR 's look good on those > >>> 2) Will this supply function unloaded while I troubleshoot it on the > >>> bench out of the unit? > >>> 3) Anyone tried independent powering the Z3801? looks like two 5V > >>> supplies and +/- 15V..not hard to to do > >>> 4) Anyone with a good supply from a parts unit want to part with same? > >>> > >>> Dave > >>> NR1DX > >>> ArtekManuals.com > >>> > >>> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@febo.com > >> To unsubscribe, go to > >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> and follow the instructions there. > >> > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
PS
paul swed
Mon, Feb 1, 2016 6:10 PM

I had the same thing happen on 1ks24361
Makes a nasty smell

On Sunday, January 31, 2016, Adrian Godwin artgodwin@gmail.com wrote:

A tant melted on one of my KS-246361 boxes (I guess they're a similar age,
possibly even the same source). The only one so far to have trouble. It
failed in such a way that it continued to smoke and load up the power
supply,

The current drawn from the no-name laptop power supply was still well
within spec, at least by the time I post-mortem'ed it. But that melted too,
and took out three levels of circuit breakers. The KS-24361 worked
perfectly after replacing the cap.

On Sun, Jan 31, 2016 at 4:42 PM, Bob Camp <kb8tq@n1k.org javascript:;>
wrote:

Hi

I had one that took out the internal fuse once every 3 to 5 years. After
replacing the fuse 3 times, the supply died. Go figure … only 15 years or
use.

Bob

On Jan 31, 2016, at 9:11 AM, paul swed <paulswedb@gmail.com

javascript:;> wrote:

As I mentioned in the another thread this week.
You can indeed replace the failed voltage with just another inverter.
Mine has run this way for 10 years. Mine was a 5 by the way. If it was

one

side of the +/-15 I would say you want to replace both.
The entire switcher can be replaced with supplies that work off of 110

so

you can get rid of the 48 volt supplies.
Good luck
Paul
WB8TSL

On Sun, Jan 31, 2016 at 12:04 AM, Alex Pummer <alex@pcscons.com

javascript:;> wrote:

power supply runs for a while, output current increases, but voltage

stays

correct -> problem is in the load circuit, load circuit working for a

while

load current increases, but it recovers after pause -> look for some
tantalum capacitor in the load circuit, they recover often  a few

times

before they finally die,
KJ6UHN
Alex

On 1/30/2016 6:21 PM, Artek Manuals wrote:

My Z3801A began to die over the last few days. It would work for an

hour

or two then produce a receiver error and communications problems.

Turn

it

off, let it sit awhile, back on and it would start the survey and

then

begin to recover. Tonight it finally went dead with high current on

the

main input . The feed supply has a 2amp current limit so I have not

blown

the two 3amp fuses

I have a schematic of the supply reverse engineered by Stefan

Hegnauere a

few years ago (thank you very much) and a limited theory of operation
description so I am not totally dead from a troubleshooting point of

view

although if one of the DC-DC converters is dead then I will likely

have to

scrap the supply

Questions

  1. Any prior experience as where to look first I checked the
    electrolytics and at first glance they seem fine, no loss of magic

smoke or

any parts obviously hot or bulging. ESR 's look good on those
2) Will this supply function unloaded while I troubleshoot it on the
bench out of the unit?
3) Anyone tried independent powering the Z3801? looks like two 5V
supplies and +/- 15V..not hard to to do
4) Anyone with a good supply from a parts unit want to part with

same?

Dave
NR1DX
ArtekManuals.com


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

I had the same thing happen on 1ks24361 Makes a nasty smell On Sunday, January 31, 2016, Adrian Godwin <artgodwin@gmail.com> wrote: > A tant melted on one of my KS-246361 boxes (I guess they're a similar age, > possibly even the same source). The only one so far to have trouble. It > failed in such a way that it continued to smoke and load up the power > supply, > > The current drawn from the no-name laptop power supply was still well > within spec, at least by the time I post-mortem'ed it. But that melted too, > and took out three levels of circuit breakers. The KS-24361 worked > perfectly after replacing the cap. > > > > > On Sun, Jan 31, 2016 at 4:42 PM, Bob Camp <kb8tq@n1k.org <javascript:;>> > wrote: > > > Hi > > > > I had one that took out the internal fuse once every 3 to 5 years. After > > replacing the fuse 3 times, the supply died. Go figure … only 15 years or > > use. > > > > Bob > > > On Jan 31, 2016, at 9:11 AM, paul swed <paulswedb@gmail.com > <javascript:;>> wrote: > > > > > > As I mentioned in the another thread this week. > > > You can indeed replace the failed voltage with just another inverter. > > > Mine has run this way for 10 years. Mine was a 5 by the way. If it was > > one > > > side of the +/-15 I would say you want to replace both. > > > The entire switcher can be replaced with supplies that work off of 110 > so > > > you can get rid of the 48 volt supplies. > > > Good luck > > > Paul > > > WB8TSL > > > > > > On Sun, Jan 31, 2016 at 12:04 AM, Alex Pummer <alex@pcscons.com > <javascript:;>> wrote: > > > > > >> power supply runs for a while, output current increases, but voltage > > stays > > >> correct -> problem is in the load circuit, load circuit working for a > > while > > >> load current increases, but it recovers after pause -> look for some > > >> tantalum capacitor in the load circuit, they recover often a few > times > > >> before they finally die, > > >> KJ6UHN > > >> Alex > > >> > > >> On 1/30/2016 6:21 PM, Artek Manuals wrote: > > >> > > >>> My Z3801A began to die over the last few days. It would work for an > > hour > > >>> or two then produce a receiver error and communications problems. > Turn > > it > > >>> off, let it sit awhile, back on and it would start the survey and > then > > >>> begin to recover. Tonight it finally went dead with high current on > the > > >>> main input . The feed supply has a 2amp current limit so I have not > > blown > > >>> the two 3amp fuses > > >>> > > >>> I have a schematic of the supply reverse engineered by Stefan > > Hegnauere a > > >>> few years ago (thank you very much) and a limited theory of operation > > >>> description so I am not totally dead from a troubleshooting point of > > view > > >>> although if one of the DC-DC converters is dead then I will likely > > have to > > >>> scrap the supply > > >>> > > >>> Questions > > >>> 1) Any prior experience as where to look first I checked the > > >>> electrolytics and at first glance they seem fine, no loss of magic > > smoke or > > >>> any parts obviously hot or bulging. ESR 's look good on those > > >>> 2) Will this supply function unloaded while I troubleshoot it on the > > >>> bench out of the unit? > > >>> 3) Anyone tried independent powering the Z3801? looks like two 5V > > >>> supplies and +/- 15V..not hard to to do > > >>> 4) Anyone with a good supply from a parts unit want to part with > same? > > >>> > > >>> Dave > > >>> NR1DX > > >>> ArtekManuals.com > > >>> > > >>> > > >> _______________________________________________ > > >> time-nuts mailing list -- time-nuts@febo.com <javascript:;> > > >> To unsubscribe, go to > > >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > >> and follow the instructions there. > > >> > > > _______________________________________________ > > > time-nuts mailing list -- time-nuts@febo.com <javascript:;> > > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > > and follow the instructions there. > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com <javascript:;> > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com <javascript:;> > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >