Hi Tom,
Can you refresh me on the topic? I didn't see any
PRS10 humps here:
http://www.leapsecond.com/museum/prs10/
We had a discussion about this, and some other folks did see it also.
Sometime around June last year or so, I need to dig up my measurements.
Just to get a clear idea of your lab set up; how is your counter
configured exactly; int or ext ref? trigger levels? 1X or 10X
attenuation, etc.
I am using the PRS10 as the external reference to the 53132A. Maybe not a
good idea? I will try another reference as well.
Also, one more datapoint: I changed the phase relationship away from 80ns,
and the problem seems to have gone away... I need to check into this further.
Maybe it's related to the absolute phase difference between the two channels?
If you can see the readings on the 53132 display, do they also
happen when your PC (RS232 or GPIB?) is not electrically
connected to the counter? Desktop or floating laptop?
RS-232, floating laptop, I get the same readings on the LED and RS232.
Is the Fury floating (wall wart) or are you using a grounded
power supply? How is the PRS10 powered? etc.
The Fury wall-wart is also floating, but the PRS10 is grounded. But I think
the periodicity of exactly 15s would indicate some sort of digital issue
rather than analog ground loop etc, wouldn't it?
Also do the 10 MHz RF outputs show this phase jump or is it
only in the 1PPS outputs?
Need to check that.
thanks again for your feedback,
bye,
Said
/tvb