time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

New NTBW50AA

AD
Arthur Dent
Tue, Sep 10, 2013 7:38 PM

Mark Sims holrum at hotmail.com
Tue Sep 10 09:03:24 EDT 2013

Again,  THE TEMPERATURE SENSOR IS NOT BROKEN!!!!!!!!!!!!!!  The firmware in
some of these units (those from NTPX modules) does NOT  read the temperature
sensor in high-res mode.

I have sold over 200 T-bolts and had to replace about 30 of their thermometer chips,
most because of the revision problem causing the 'stepped' temp plot, but some
because of chip failure. If this were the typical low res problem the yellow trace
would show the characteristic uniform staircase steps, not the random jumps to
some common  value that it never exceeds. If you look closely at the plot you'll
see many very small changes that are far smaller than any of the steps would be.
Also there are major portions of the trace where it is absolutely flat and I doubt that
this is really happening. I'll stick with broken but maybe someone else with another
NTBW50AA can run LH and show what their graph looks like.

-Arthur

>Mark Sims holrum at hotmail.com >Tue Sep 10 09:03:24 EDT 2013 > >Again,  THE TEMPERATURE SENSOR IS NOT BROKEN!!!!!!!!!!!!!!  The firmware in >some of these units (those from NTPX modules) does NOT  read the temperature >sensor in high-res mode. I have sold over 200 T-bolts and had to replace about 30 of their thermometer chips, most because of the revision problem causing the 'stepped' temp plot, but some because of chip failure. If this were the typical low res problem the yellow trace would show the characteristic uniform staircase steps, not the random jumps to some common  value that it never exceeds. If you look closely at the plot you'll see many very small changes that are far smaller than any of the steps would be. Also there are major portions of the trace where it is absolutely flat and I doubt that this is really happening. I'll stick with broken but maybe someone else with another NTBW50AA can run LH and show what their graph looks like. -Arthur
Q
quartz55
Tue, Sep 10, 2013 9:31 PM

Arthur,

Can you tell me where and what the chip is and if there's any way to test it on the board?  I've got a 30 day $ back deal on this thing.

Dave

Arthur, Can you tell me where and what the chip is and if there's any way to test it on the board? I've got a 30 day $ back deal on this thing. Dave
BC
Bob Camp
Tue, Sep 10, 2013 9:56 PM

Hi

I have quite a few of these and they all produce temperature plots that look like proper TBolts. I also have TBolts that don't read the sensor in high resolution mode. I have a couple TBolts with broken temperature sensor chips. Assuming the room temperature is varying several degrees C up and down over the time shown, those plots most closely resemble a TBolt with a blown chip.

Bob

On Sep 10, 2013, at 3:38 PM, Arthur Dent golgarfrincham@yahoo.com wrote:

Mark Sims holrum at hotmail.com
Tue Sep 10 09:03:24 EDT 2013

Again,  THE TEMPERATURE SENSOR IS NOT BROKEN!!!!!!!!!!!!!!  The firmware in
some of these units (those from NTPX modules) does NOT  read the temperature
sensor in high-res mode.

I have sold over 200 T-bolts and had to replace about 30 of their thermometer chips,
most because of the revision problem causing the 'stepped' temp plot, but some
because of chip failure. If this were the typical low res problem the yellow trace
would show the characteristic uniform staircase steps, not the random jumps to
some common  value that it never exceeds. If you look closely at the plot you'll
see many very small changes that are far smaller than any of the steps would be.
Also there are major portions of the trace where it is absolutely flat and I doubt that
this is really happening. I'll stick with broken but maybe someone else with another
NTBW50AA can run LH and show what their graph looks like.

-Arthur


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi I have quite a few of these and they all produce temperature plots that look like proper TBolts. I also have TBolts that don't read the sensor in high resolution mode. I have a couple TBolts with broken temperature sensor chips. Assuming the room temperature is varying several degrees C up and down over the time shown, those plots most closely resemble a TBolt with a blown chip. Bob On Sep 10, 2013, at 3:38 PM, Arthur Dent <golgarfrincham@yahoo.com> wrote: >> Mark Sims holrum at hotmail.com >> Tue Sep 10 09:03:24 EDT 2013 >> >> Again, THE TEMPERATURE SENSOR IS NOT BROKEN!!!!!!!!!!!!!! The firmware in >> some of these units (those from NTPX modules) does NOT read the temperature >> sensor in high-res mode. > > I have sold over 200 T-bolts and had to replace about 30 of their thermometer chips, > most because of the revision problem causing the 'stepped' temp plot, but some > because of chip failure. If this were the typical low res problem the yellow trace > would show the characteristic uniform staircase steps, not the random jumps to > some common value that it never exceeds. If you look closely at the plot you'll > see many very small changes that are far smaller than any of the steps would be. > Also there are major portions of the trace where it is absolutely flat and I doubt that > this is really happening. I'll stick with broken but maybe someone else with another > NTBW50AA can run LH and show what their graph looks like. > > -Arthur > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BC
Bob Camp
Tue, Sep 10, 2013 9:58 PM

Hi

Simple way to check it:

Let the temperature in the room move over a 10 C range (open a window …). If the temperature trace does not move the chip isn't doing what it should.

Bob

On Sep 10, 2013, at 5:31 PM, quartz55 quartz55@hughes.net wrote:

Arthur,

Can you tell me where and what the chip is and if there's any way to test it on the board?  I've got a 30 day $ back deal on this thing.

Dave


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Simple way to check it: Let the temperature in the room move over a 10 C range (open a window …). If the temperature trace does not move the chip isn't doing what it should. Bob On Sep 10, 2013, at 5:31 PM, quartz55 <quartz55@hughes.net> wrote: > Arthur, > > Can you tell me where and what the chip is and if there's any way to test it on the board? I've got a 30 day $ back deal on this thing. > > Dave > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
Q
quartz55
Wed, Sep 11, 2013 12:43 AM

Be a lot easier just to blow the heat gun on it.  What is this sensor measuring, ambient temperature in the room of something on the board, like the OCXO or what?  It only seems to vary from about 37.7 to 41.7 and it always seems to change in 1° increments.  Is it going to affect how the GPSDO works?

Dave

Be a lot easier just to blow the heat gun on it. What is this sensor measuring, ambient temperature in the room of something on the board, like the OCXO or what? It only seems to vary from about 37.7 to 41.7 and it always seems to change in 1° increments. Is it going to affect how the GPSDO works? Dave
BC
Bob Camp
Wed, Sep 11, 2013 12:52 AM

Hi

It should be measuring ambient temperature. The GPSDO probably uses it as part of the holdover process. A good blast of room temperature air (heat gun without the heater on) should drop the sensor 5 to 10 degrees.

Bob

On Sep 10, 2013, at 8:43 PM, quartz55 quartz55@hughes.net wrote:

Be a lot easier just to blow the heat gun on it.  What is this sensor measuring, ambient temperature in the room of something on the board, like the OCXO or what?  It only seems to vary from about 37.7 to 41.7 and it always seems to change in 1° increments.  Is it going to affect how the GPSDO works?

Dave


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi It should be measuring ambient temperature. The GPSDO probably uses it as part of the holdover process. A good blast of room temperature air (heat gun without the heater on) *should* drop the sensor 5 to 10 degrees. Bob On Sep 10, 2013, at 8:43 PM, quartz55 <quartz55@hughes.net> wrote: > Be a lot easier just to blow the heat gun on it. What is this sensor measuring, ambient temperature in the room of something on the board, like the OCXO or what? It only seems to vary from about 37.7 to 41.7 and it always seems to change in 1° increments. Is it going to affect how the GPSDO works? > > Dave > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
Q
quartz55
Wed, Sep 11, 2013 2:02 AM

Well, here's a shot with me putting a fan under the bottom of the unit and you can see the temp immediately start to drop from 39.7 down to 29.7 and then I take the fan away and it goes back up to 37.7 and now I've let the bottom breathe which I had blocked off by letting it sit on the bottom.  I would say the temp sensor is working, but still why those jumps?  I'll let it run with the bottom/top open and see what happens then.  But it was doing these jumps with the cover off before as I remember.  Wonder if I should have a bit of forced ventilation in this thing?
http://s251.photobucket.com/user/DogTi/media/time/temptest_zps4eda29a5.jpg.html?sort=3&o=0

Dave

Well, here's a shot with me putting a fan under the bottom of the unit and you can see the temp immediately start to drop from 39.7 down to 29.7 and then I take the fan away and it goes back up to 37.7 and now I've let the bottom breathe which I had blocked off by letting it sit on the bottom. I would say the temp sensor is working, but still why those jumps? I'll let it run with the bottom/top open and see what happens then. But it was doing these jumps with the cover off before as I remember. Wonder if I should have a bit of forced ventilation in this thing? http://s251.photobucket.com/user/DogTi/media/time/temptest_zps4eda29a5.jpg.html?sort=3&o=0 Dave
BC
Bob Camp
Wed, Sep 11, 2013 11:07 AM

Hi

Ok, you have an unusually well controlled environment in your shack.

The pops are the temperature just barely making it to a 1 C excursion. Think of your temperature as a sort of sine wave. The pops are the peaks of the sine wave.

Bob

On Sep 10, 2013, at 10:02 PM, quartz55 quartz55@hughes.net wrote:

Well, here's a shot with me putting a fan under the bottom of the unit and you can see the temp immediately start to drop from 39.7 down to 29.7 and then I take the fan away and it goes back up to 37.7 and now I've let the bottom breathe which I had blocked off by letting it sit on the bottom.  I would say the temp sensor is working, but still why those jumps?  I'll let it run with the bottom/top open and see what happens then.  But it was doing these jumps with the cover off before as I remember.  Wonder if I should have a bit of forced ventilation in this thing?
http://s251.photobucket.com/user/DogTi/media/time/temptest_zps4eda29a5.jpg.html?sort=3&o=0

Dave


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Ok, you have an unusually well controlled environment in your shack. The pops are the temperature just barely making it to a 1 C excursion. Think of your temperature as a sort of sine wave. The pops are the peaks of the sine wave. Bob On Sep 10, 2013, at 10:02 PM, quartz55 <quartz55@hughes.net> wrote: > Well, here's a shot with me putting a fan under the bottom of the unit and you can see the temp immediately start to drop from 39.7 down to 29.7 and then I take the fan away and it goes back up to 37.7 and now I've let the bottom breathe which I had blocked off by letting it sit on the bottom. I would say the temp sensor is working, but still why those jumps? I'll let it run with the bottom/top open and see what happens then. But it was doing these jumps with the cover off before as I remember. Wonder if I should have a bit of forced ventilation in this thing? > http://s251.photobucket.com/user/DogTi/media/time/temptest_zps4eda29a5.jpg.html?sort=3&o=0 > > Dave > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
Q
quartz55
Wed, Sep 11, 2013 12:42 PM

I set the temp scale to 250m°C/div last night and it remained steady for quit a while, but then it started the jumping again.  It seems to jump on 1° increments, like from 37.750 to 38.749 in less than 1 minute, but there are step variations during the 1°  movements.  So what you're saying is it won't affect the freq stab or adev?  And yes the room temp is quite stable, it's about 1/2 underground to 3 sides and the A/C runs most of the time this time of year.  I guess it's not worth hounding the seller about it if the others he has are just like it.  It's just a matter of if I want to replace the chip.  Looks like they're up to DS1620S by now.

I'm not going to worry about it for now.  But I am going to move the antenna over to where it sees the sky without the trees this morning before it gets to 90°.

Dave

I set the temp scale to 250m°C/div last night and it remained steady for quit a while, but then it started the jumping again. It seems to jump on 1° increments, like from 37.750 to 38.749 in less than 1 minute, but there are step variations during the 1° movements. So what you're saying is it won't affect the freq stab or adev? And yes the room temp is quite stable, it's about 1/2 underground to 3 sides and the A/C runs most of the time this time of year. I guess it's not worth hounding the seller about it if the others he has are just like it. It's just a matter of if I want to replace the chip. Looks like they're up to DS1620S by now. I'm not going to worry about it for now. But I am going to move the antenna over to where it sees the sky without the trees this morning before it gets to 90°. Dave
CS
Charles Steinmetz
Wed, Sep 11, 2013 1:55 PM

Dave wrote:

I set the temp scale to 250m°C/div last night
and it remained steady for quit a while, but
then it started the jumping again.  It seems to
jump on 1° increments, like from 37.750 to
38.749 in less than 1 minute, but there are step
variations during the 1°  movements.

I do not know whether LH just displays every
"n-th" sample to produce long-time-scale plots,
or whether it averages "n" samples to get the
displayed value.  If it averages, each pixel
would include data from more than one sample,
which would produce intermediate values that the
GPSDO did not actually report.  You may want to
set the time scale so you can see each one-second
sample (I think one minute per division will do
this).  At that scale, you may find that all of
the samples are in one degree increments.  In
other words, the appearance of intermediate
values may be an artifact of display averaging --
the actual data from the GPSDO may be strictly in
1v increments.  Mark could say for sure.

So what you're saying is it won't affect the freq stab or adev?

We (the time-nuts community) do not have hard
data about how the temperature data is used in
the Trimble GPSDOs, but it appears that it is not
used during locked operation except to "teach"
the unit about the OCXO's temperature
coefficient.  What it has "learned" is applied
during holdover in an attempt to improve holdover
stability.  I doubt many time nuts would make
critical measurements while their GPSDO was
unlocked, so there is probably no practical
difference as far as we are concerned.

Best regards,

Charles

Dave wrote: >I set the temp scale to 250m°C/div last night >and it remained steady for quit a while, but >then it started the jumping again. It seems to >jump on 1° increments, like from 37.750 to >38.749 in less than 1 minute, but there are step >variations during the 1° movements. I do not know whether LH just displays every "n-th" sample to produce long-time-scale plots, or whether it averages "n" samples to get the displayed value. If it averages, each pixel would include data from more than one sample, which would produce intermediate values that the GPSDO did not actually report. You may want to set the time scale so you can see each one-second sample (I think one minute per division will do this). At that scale, you may find that all of the samples are in one degree increments. In other words, the appearance of intermediate values may be an artifact of display averaging -- the actual data from the GPSDO may be strictly in 1v increments. Mark could say for sure. >So what you're saying is it won't affect the freq stab or adev? We (the time-nuts community) do not have hard data about how the temperature data is used in the Trimble GPSDOs, but it appears that it is not used during locked operation except to "teach" the unit about the OCXO's temperature coefficient. What it has "learned" is applied during holdover in an attempt to improve holdover stability. I doubt many time nuts would make critical measurements while their GPSDO was unlocked, so there is probably no practical difference as far as we are concerned. Best regards, Charles
Q
quartz55
Wed, Sep 11, 2013 2:17 PM

Thanks Charles,

I'm not certain myself it it's the temp chip or the way LH handles the information.  It doesn't matter like you say, there's no use doing any measurements with the unit unlocked.

Otherwise here is a trace while I moved the antenna.  Any comments on how it reacted in holdover?
http://i251.photobucket.com/albums/gg287/DogTi/time/holdover_zps35d7e81b.jpg
And here is where the antenna is now, this is looking straight north, so the sky behind me is nearly clear to the horizon and straight up isn't bad either.
http://i251.photobucket.com/albums/gg287/DogTi/time/gpsant_zps2963365a.jpg

Dave
N3DT

Thanks Charles, I'm not certain myself it it's the temp chip or the way LH handles the information. It doesn't matter like you say, there's no use doing any measurements with the unit unlocked. Otherwise here is a trace while I moved the antenna. Any comments on how it reacted in holdover? http://i251.photobucket.com/albums/gg287/DogTi/time/holdover_zps35d7e81b.jpg And here is where the antenna is now, this is looking straight north, so the sky behind me is nearly clear to the horizon and straight up isn't bad either. http://i251.photobucket.com/albums/gg287/DogTi/time/gpsant_zps2963365a.jpg Dave N3DT
M
mc235960
Wed, Sep 11, 2013 3:42 PM

I am tempted to use the monty python exhortation "Start again"

If I am reading your screen shots correctly the 01:51:57 UTC shot is before the antenna move and the 14:02:03 one after the move.
HOWEVER, both shots show the same saved position. That is certainly not going to be optimal. You should let it survey again.

Mike

Le 11 sept. 2013 à 16:17, quartz55 a écrit :

Thanks Charles,

I'm not certain myself it it's the temp chip or the way LH handles the information.  It doesn't matter like you say, there's no use doing any measurements with the unit unlocked.

Otherwise here is a trace while I moved the antenna.  Any comments on how it reacted in holdover?
http://i251.photobucket.com/albums/gg287/DogTi/time/holdover_zps35d7e81b.jpg
And here is where the antenna is now, this is looking straight north, so the sky behind me is nearly clear to the horizon and straight up isn't bad either.
http://i251.photobucket.com/albums/gg287/DogTi/time/gpsant_zps2963365a.jpg

Dave
N3DT


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

I am tempted to use the monty python exhortation "Start again" If I am reading your screen shots correctly the 01:51:57 UTC shot is before the antenna move and the 14:02:03 one after the move. HOWEVER, both shots show the same saved position. That is certainly not going to be optimal. You should let it survey again. Mike Le 11 sept. 2013 à 16:17, quartz55 a écrit : > Thanks Charles, > > I'm not certain myself it it's the temp chip or the way LH handles the information. It doesn't matter like you say, there's no use doing any measurements with the unit unlocked. > > Otherwise here is a trace while I moved the antenna. Any comments on how it reacted in holdover? > http://i251.photobucket.com/albums/gg287/DogTi/time/holdover_zps35d7e81b.jpg > And here is where the antenna is now, this is looking straight north, so the sky behind me is nearly clear to the horizon and straight up isn't bad either. > http://i251.photobucket.com/albums/gg287/DogTi/time/gpsant_zps2963365a.jpg > > Dave > N3DT > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
W
WarrenS
Wed, Sep 11, 2013 5:10 PM

Dave

With your GPS Antenna sitting underneath all those multipath reflectors (the
other antenna's above it),
That is far from optimal from a time-nut standpoint if you are trying to get
the best performance possible.
Should strive for no metal or reflective surfaces above the GPS antenna in
any direction.

ws


----- Original Message -----
From: "quartz55"
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Subject: Re: [time-nuts] New NTBW50AA

Thanks Charles,

I'm not certain myself it it's the temp chip or the way LH handles the
information.  It doesn't matter like you say, there's no use doing any
measurements with the unit unlocked.

Otherwise here is a trace while I moved the antenna.  Any comments on how it
reacted in holdover?
http://i251.photobucket.com/albums/gg287/DogTi/time/holdover_zps35d7e81b.jpg
And here is where the antenna is now, this is looking straight north, so the
sky behind me is nearly clear to the horizon and straight up isn't bad
either.
http://i251.photobucket.com/albums/gg287/DogTi/time/gpsant_zps2963365a.jpg

Dave
N3DT

Dave With your GPS Antenna sitting underneath all those multipath reflectors (the other antenna's above it), That is far from optimal from a time-nut standpoint if you are trying to get the best performance possible. Should strive for no metal or reflective surfaces above the GPS antenna in any direction. ws ********************* ----- Original Message ----- From: "quartz55" To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com> Subject: Re: [time-nuts] New NTBW50AA Thanks Charles, I'm not certain myself it it's the temp chip or the way LH handles the information. It doesn't matter like you say, there's no use doing any measurements with the unit unlocked. Otherwise here is a trace while I moved the antenna. Any comments on how it reacted in holdover? http://i251.photobucket.com/albums/gg287/DogTi/time/holdover_zps35d7e81b.jpg And here is where the antenna is now, this is looking straight north, so the sky behind me is nearly clear to the horizon and straight up isn't bad either. http://i251.photobucket.com/albums/gg287/DogTi/time/gpsant_zps2963365a.jpg Dave N3DT
BS
Bob Stewart
Wed, Sep 11, 2013 5:21 PM

I can testify to Warren's comments. I've got an Adafruit Breakout Board that I hook up to my computer from time to time, and it marches all around the neighborhood on the map looking for my house.  =)  Eventually my timing antenna will go up in the attic, just under the peak.

Bob - AE6RV


From: WarrenS warrensjmail-one@yahoo.com
To: Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Wednesday, September 11, 2013 12:10 PM
Subject: Re: [time-nuts] New NTBW50AA

Dave

With your GPS Antenna sitting underneath all those multipath reflectors (the other antenna's above it),
That is far from optimal from a time-nut standpoint if you are trying to get the best performance possible.
Should strive for no metal or reflective surfaces above the GPS antenna in any direction.

ws


----- Original Message ----- From: "quartz55"
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Subject: Re: [time-nuts] New NTBW50AA

Thanks Charles,

I'm not certain myself it it's the temp chip or the way LH handles the information.  It doesn't matter like you say, there's no use doing any measurements with the unit unlocked.

Otherwise here is a trace while I moved the antenna.  Any comments on how it reacted in holdover?
http://i251.photobucket.com/albums/gg287/DogTi/time/holdover_zps35d7e81b.jpg
And here is where the antenna is now, this is looking straight north, so the sky behind me is nearly clear to the horizon and straight up isn't bad either.
http://i251.photobucket.com/albums/gg287/DogTi/time/gpsant_zps2963365a.jpg

Dave
N3DT


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

I can testify to Warren's comments. I've got an Adafruit Breakout Board that I hook up to my computer from time to time, and it marches all around the neighborhood on the map looking for my house.  =)  Eventually my timing antenna will go up in the attic, just under the peak. Bob - AE6RV >________________________________ > From: WarrenS <warrensjmail-one@yahoo.com> >To: Discussion of precise time and frequency measurement <time-nuts@febo.com> >Sent: Wednesday, September 11, 2013 12:10 PM >Subject: Re: [time-nuts] New NTBW50AA > > > >Dave > >With your GPS Antenna sitting underneath all those multipath reflectors (the other antenna's above it), >That is far from optimal from a time-nut standpoint if you are trying to get the best performance possible. >Should strive for no metal or reflective surfaces above the GPS antenna in any direction. > >ws > >********************* > >----- Original Message ----- From: "quartz55" >To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com> >Subject: Re: [time-nuts] New NTBW50AA > > >Thanks Charles, > >I'm not certain myself it it's the temp chip or the way LH handles the information.  It doesn't matter like you say, there's no use doing any measurements with the unit unlocked. > >Otherwise here is a trace while I moved the antenna.  Any comments on how it reacted in holdover? >http://i251.photobucket.com/albums/gg287/DogTi/time/holdover_zps35d7e81b.jpg >And here is where the antenna is now, this is looking straight north, so the sky behind me is nearly clear to the horizon and straight up isn't bad either. >http://i251.photobucket.com/albums/gg287/DogTi/time/gpsant_zps2963365a.jpg > >Dave >N3DT > >_______________________________________________ >time-nuts mailing list -- time-nuts@febo.com >To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >and follow the instructions there. > > >
Q
quartz55
Wed, Sep 11, 2013 5:49 PM

The lat lon did change when I did the survey.  On google, the lat lon is real close, a foot or so for the before position and after also.  It certainly doesn't march around the neighborhood looking, the last survey must have only taken a few minutes.  As far as antenna position, I guess maybe I can put it over on the hughes dish, but it's closer to the trees there.  How far away from those beams do I need to get, so they're not in the el mask?  Maybe I could put it on one of the roof vents but the coax will then drag across the shingles.  I'll have to think about this.  On the deck is good, but the wifey wouldn't go for that and it would be lots lower.  Maybe if I paint it flat black she won't see it.

You can see where google reports it here and the arrow is more where it is, so maybe it's 4-5' off.
http://i251.photobucket.com/albums/gg287/DogTi/time/roof_zps5eccff6a.jpg

Dave

The lat lon did change when I did the survey. On google, the lat lon is real close, a foot or so for the before position and after also. It certainly doesn't march around the neighborhood looking, the last survey must have only taken a few minutes. As far as antenna position, I guess maybe I can put it over on the hughes dish, but it's closer to the trees there. How far away from those beams do I need to get, so they're not in the el mask? Maybe I could put it on one of the roof vents but the coax will then drag across the shingles. I'll have to think about this. On the deck is good, but the wifey wouldn't go for that and it would be lots lower. Maybe if I paint it flat black she won't see it. You can see where google reports it here and the arrow is more where it is, so maybe it's 4-5' off. http://i251.photobucket.com/albums/gg287/DogTi/time/roof_zps5eccff6a.jpg Dave
Q
quartz55
Wed, Sep 11, 2013 5:50 PM

Oh, the XRef from VK3HZ is at the PO, I'm going to go get it.

Dave

Oh, the XRef from VK3HZ is at the PO, I'm going to go get it. Dave
BC
Bob Camp
Wed, Sep 11, 2013 10:49 PM

Hi

Ideally you would like the position error to be under 10" for the device to work at it's best. That's probably going to equate to a couple of inches of error in X, Y, and Z.  Since GPS has cyclical errors, you need an average over a few days to get to that level.

Bob

On Sep 11, 2013, at 1:49 PM, quartz55 quartz55@hughes.net wrote:

The lat lon did change when I did the survey.  On google, the lat lon is real close, a foot or so for the before position and after also.  It certainly doesn't march around the neighborhood looking, the last survey must have only taken a few minutes.  As far as antenna position, I guess maybe I can put it over on the hughes dish, but it's closer to the trees there.  How far away from those beams do I need to get, so they're not in the el mask?  Maybe I could put it on one of the roof vents but the coax will then drag across the shingles.  I'll have to think about this.  On the deck is good, but the wifey wouldn't go for that and it would be lots lower.  Maybe if I paint it flat black she won't see it.

You can see where google reports it here and the arrow is more where it is, so maybe it's 4-5' off.
http://i251.photobucket.com/albums/gg287/DogTi/time/roof_zps5eccff6a.jpg

Dave


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Ideally you would like the position error to be under 10" for the device to work at it's best. That's probably going to equate to a couple of inches of error in X, Y, and Z. Since GPS has cyclical errors, you need an average over a few days to get to that level. Bob On Sep 11, 2013, at 1:49 PM, quartz55 <quartz55@hughes.net> wrote: > The lat lon did change when I did the survey. On google, the lat lon is real close, a foot or so for the before position and after also. It certainly doesn't march around the neighborhood looking, the last survey must have only taken a few minutes. As far as antenna position, I guess maybe I can put it over on the hughes dish, but it's closer to the trees there. How far away from those beams do I need to get, so they're not in the el mask? Maybe I could put it on one of the roof vents but the coax will then drag across the shingles. I'll have to think about this. On the deck is good, but the wifey wouldn't go for that and it would be lots lower. Maybe if I paint it flat black she won't see it. > > You can see where google reports it here and the arrow is more where it is, so maybe it's 4-5' off. > http://i251.photobucket.com/albums/gg287/DogTi/time/roof_zps5eccff6a.jpg > > Dave > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
Q
quartz55
Thu, Sep 12, 2013 1:48 AM

I ordered a couple of the temp chips, but it's going to be low on the list of priorities.

Tomorrow I'll add another 30' of cable and mount the antenna on the kitchen vent which is right on the south end of the house above the deck, lower but southeast from the beams by 25' or so.  I think that's about as good as I can do unless I put up a tower to get above everything and that's not going to happen.

I broke a chip inductor in the TS-2000 trying to install the XRef today, so I have to wait a few days until I get some new parts in.  I can do a 24 hour precision survey in the meantime.  I had to get the headband magnifier out to work on this thing, the board is only 1.25" x 0 .4".  The chip inductor I broke is only about a 2mm cube.  I got it off the 2000 board OK, I broke the track off the chip trying to solder a wire on one end.  Next time I'm going to solder it and then glue it.
http://i251.photobucket.com/albums/gg287/DogTi/time/XRef_zps74f23696.jpg

I've got about 2 acres of grass to cut and need to put the sickle bar on the JD and go out and kill a few stink bugs in this heat.

Dave
N3DT

I ordered a couple of the temp chips, but it's going to be low on the list of priorities. Tomorrow I'll add another 30' of cable and mount the antenna on the kitchen vent which is right on the south end of the house above the deck, lower but southeast from the beams by 25' or so. I think that's about as good as I can do unless I put up a tower to get above everything and that's not going to happen. I broke a chip inductor in the TS-2000 trying to install the XRef today, so I have to wait a few days until I get some new parts in. I can do a 24 hour precision survey in the meantime. I had to get the headband magnifier out to work on this thing, the board is only 1.25" x 0 .4". The chip inductor I broke is only about a 2mm cube. I got it off the 2000 board OK, I broke the track off the chip trying to solder a wire on one end. Next time I'm going to solder it and then glue it. http://i251.photobucket.com/albums/gg287/DogTi/time/XRef_zps74f23696.jpg I've got about 2 acres of grass to cut and need to put the sickle bar on the JD and go out and kill a few stink bugs in this heat. Dave N3DT
BC
Bob Camp
Thu, Sep 12, 2013 11:05 AM

Hi

As you do these surveys, write down the results. You probably will do several surveys on each antenna location before you are done. Looking at the variation between the results will give you an idea of what's going on.

Bob

On Sep 11, 2013, at 9:48 PM, quartz55 quartz55@hughes.net wrote:

I ordered a couple of the temp chips, but it's going to be low on the list of priorities.

Tomorrow I'll add another 30' of cable and mount the antenna on the kitchen vent which is right on the south end of the house above the deck, lower but southeast from the beams by 25' or so.  I think that's about as good as I can do unless I put up a tower to get above everything and that's not going to happen.

I broke a chip inductor in the TS-2000 trying to install the XRef today, so I have to wait a few days until I get some new parts in.  I can do a 24 hour precision survey in the meantime.  I had to get the headband magnifier out to work on this thing, the board is only 1.25" x 0 .4".  The chip inductor I broke is only about a 2mm cube.  I got it off the 2000 board OK, I broke the track off the chip trying to solder a wire on one end.  Next time I'm going to solder it and then glue it.
http://i251.photobucket.com/albums/gg287/DogTi/time/XRef_zps74f23696.jpg

I've got about 2 acres of grass to cut and need to put the sickle bar on the JD and go out and kill a few stink bugs in this heat.

Dave
N3DT


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi As you do these surveys, write down the results. You probably will do several surveys on each antenna location before you are done. Looking at the variation between the results will give you an idea of what's going on. Bob On Sep 11, 2013, at 9:48 PM, quartz55 <quartz55@hughes.net> wrote: > I ordered a couple of the temp chips, but it's going to be low on the list of priorities. > > Tomorrow I'll add another 30' of cable and mount the antenna on the kitchen vent which is right on the south end of the house above the deck, lower but southeast from the beams by 25' or so. I think that's about as good as I can do unless I put up a tower to get above everything and that's not going to happen. > > I broke a chip inductor in the TS-2000 trying to install the XRef today, so I have to wait a few days until I get some new parts in. I can do a 24 hour precision survey in the meantime. I had to get the headband magnifier out to work on this thing, the board is only 1.25" x 0 .4". The chip inductor I broke is only about a 2mm cube. I got it off the 2000 board OK, I broke the track off the chip trying to solder a wire on one end. Next time I'm going to solder it and then glue it. > http://i251.photobucket.com/albums/gg287/DogTi/time/XRef_zps74f23696.jpg > > I've got about 2 acres of grass to cut and need to put the sickle bar on the JD and go out and kill a few stink bugs in this heat. > > Dave > N3DT > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
Q
quartz55
Thu, Sep 12, 2013 1:51 PM

Will too long a cable run or too many connectors (2) result in "antenna open"?  When I added the last 50' of cable that's what I got.  I measure 5V at the end of the cable run though.  Maybe some de-ox in the connectors will help?  I can also go get a single long run instead of piecing it together.  I'll try a few things before heading into town.

Dave

Will too long a cable run or too many connectors (2) result in "antenna open"? When I added the last 50' of cable that's what I got. I measure 5V at the end of the cable run though. Maybe some de-ox in the connectors will help? I can also go get a single long run instead of piecing it together. I'll try a few things before heading into town. Dave
Q
quartz55
Thu, Sep 12, 2013 2:10 PM

OK, I cleaned up the connectors and it's working again at the new location. As soon as I get into town, I'll get some new coax.

Dave

OK, I cleaned up the connectors and it's working again at the new location. As soon as I get into town, I'll get some new coax. Dave
BC
Bob Camp
Thu, Sep 12, 2013 9:59 PM

Hi

Looks like RG-6 Quad Shield is about $50 for a 500' spool at the local big box stores.

Bob

On Sep 12, 2013, at 10:10 AM, quartz55 quartz55@hughes.net wrote:

OK, I cleaned up the connectors and it's working again at the new location. As soon as I get into town, I'll get some new coax.

Dave


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Looks like RG-6 Quad Shield is about $50 for a 500' spool at the local big box stores. Bob On Sep 12, 2013, at 10:10 AM, quartz55 <quartz55@hughes.net> wrote: > OK, I cleaned up the connectors and it's working again at the new location. As soon as I get into town, I'll get some new coax. > > Dave > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
Q
quartz55
Fri, Sep 13, 2013 1:50 AM

Well, I was in HD tonight and 100' is $25.  Otherwise they had a roll of 500' for around $50.  Nothing in between.  You have to realize nothing out here in the boonies comes easy.  I only need maybe 150' and actually the 3 pieces of 50' I have are not that old.  We'll see what happens.  Right now it seems to be working  a little better with the antenna on the kitchen vent.  I don't know what I'd do with 500'.

Dave
----- Original Message -----
From: Bob Camp
To: Discussion of precise time and frequency measurement
Sent: Thursday, September 12, 2013 5:59 PM
Subject: Re: [time-nuts] New NTBW50AA

Hi

Looks like RG-6 Quad Shield is about $50 for a 500' spool at the local big box stores.

Bob

Well, I was in HD tonight and 100' is $25. Otherwise they had a roll of 500' for around $50. Nothing in between. You have to realize nothing out here in the boonies comes easy. I only need maybe 150' and actually the 3 pieces of 50' I have are not that old. We'll see what happens. Right now it seems to be working a little better with the antenna on the kitchen vent. I don't know what I'd do with 500'. Dave ----- Original Message ----- From: Bob Camp To: Discussion of precise time and frequency measurement Sent: Thursday, September 12, 2013 5:59 PM Subject: Re: [time-nuts] New NTBW50AA Hi Looks like RG-6 Quad Shield is about $50 for a 500' spool at the local big box stores. Bob
Q
quartz55
Fri, Sep 13, 2013 1:59 AM

The LH OSCdisplay is measuring 'parts per trillion/div' of what?  Sorry for the basic question, but I have not found it.  At least I figured out it was trillion, not thousand.

Dave

The LH OSCdisplay is measuring 'parts per trillion/div' of what? Sorry for the basic question, but I have not found it. At least I figured out it was trillion, not thousand. Dave
BC
Bob Camp
Fri, Sep 13, 2013 11:04 AM

Hi

Unless you have it shipped in, that's pretty much the situation everywhere. I have several lifetime supply spools of this and that. Some of them I bought new. Others I picked up at garage sales, still with a lifetime supply of cable on the spool. Oddly enough some were replacements for other "lifetime" supplies that got used up….

Bob

On Sep 12, 2013, at 9:50 PM, quartz55 quartz55@hughes.net wrote:

Well, I was in HD tonight and 100' is $25.  Otherwise they had a roll of 500' for around $50.  Nothing in between.  You have to realize nothing out here in the boonies comes easy.  I only need maybe 150' and actually the 3 pieces of 50' I have are not that old.  We'll see what happens.  Right now it seems to be working  a little better with the antenna on the kitchen vent.  I don't know what I'd do with 500'.

Dave
----- Original Message -----
From: Bob Camp
To: Discussion of precise time and frequency measurement
Sent: Thursday, September 12, 2013 5:59 PM
Subject: Re: [time-nuts] New NTBW50AA

Hi

Looks like RG-6 Quad Shield is about $50 for a 500' spool at the local big box stores.

Bob


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Unless you have it shipped in, that's pretty much the situation everywhere. I have several lifetime supply spools of this and that. Some of them I bought new. Others I picked up at garage sales, still with a lifetime supply of cable on the spool. Oddly enough some were replacements for other "lifetime" supplies that got used up…. Bob On Sep 12, 2013, at 9:50 PM, quartz55 <quartz55@hughes.net> wrote: > Well, I was in HD tonight and 100' is $25. Otherwise they had a roll of 500' for around $50. Nothing in between. You have to realize nothing out here in the boonies comes easy. I only need maybe 150' and actually the 3 pieces of 50' I have are not that old. We'll see what happens. Right now it seems to be working a little better with the antenna on the kitchen vent. I don't know what I'd do with 500'. > > Dave > ----- Original Message ----- > From: Bob Camp > To: Discussion of precise time and frequency measurement > Sent: Thursday, September 12, 2013 5:59 PM > Subject: Re: [time-nuts] New NTBW50AA > > > Hi > > Looks like RG-6 Quad Shield is about $50 for a 500' spool at the local big box stores. > > Bob > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
Q
quartz55
Fri, Sep 13, 2013 12:59 PM

Yeah, that's what I'm worried about, I've got enough lifetime supplies of junk around here now.
Oddly enough some were replacements for other "lifetime" supplies that got used up….

Bob

Yeah, that's what I'm worried about, I've got enough lifetime supplies of junk around here now. Oddly enough some were replacements for other "lifetime" supplies that got used up…. Bob
Q
quartz55
Sat, Sep 14, 2013 4:04 AM

I understand it means parts per trillion (10^12), but can someone tell me what the trace is measuring?  Is it an indication of how far off frequency the osc is at incremental times, relative to what the unit thinks is correct by way of PPS (over time)?  Or point me to reading material please.  I'm assuming the object for precise frequency is to get this trace as stable and the smallest variation possible?

Dave

I understand it means parts per trillion (10^12), but can someone tell me what the trace is measuring? Is it an indication of how far off frequency the osc is at incremental times, relative to what the unit thinks is correct by way of PPS (over time)? Or point me to reading material please. I'm assuming the object for precise frequency is to get this trace as stable and the smallest variation possible? Dave
Q
quartz55
Sat, Sep 14, 2013 4:45 PM

Today I moved the antenna to above the beams.  This is temporary because it's in the way of the beams turning.  I guess I need to get someone out here to help me install it on top of the beam mast.  I don't feel confident to do it myself anymore.  I suppose it's not going to hurt anything if once in a while the antenna gets rotated?  Just not while I'm doing something with the Nortel.  It's pretty much in the clear and as high as I can get it.  It sees clear sky especially to the south except for one tree and the antenna is probably within 10' of that tree's height.  There are still trees to the east that I just can't do much about.  The ones to the west are about 100' away and the antenna shouldn't see them much.

http://i251.photobucket.com/albums/gg287/DogTi/time/abovebeams_zpsbf8474c1.jpg

So the Osc ppt is the offset from what the Nortel thinks is exactly 10MHz as I read it on page 117 of the ThunderBolt instructions (thanks Tom).  How do I tell how accurate the Nortel thinks exactly 10MHz is, or can I assume it's right?  If right is the case, then if I get a spread of around 300ppt, that means I'm always within 300x10^-9 Hz of 10MHz or .0003Hz at 1GHz?  Or is my thinking off base and did I miss a decimal somewhere?  I suppose the ppt spread is pretty much a function of how stable the osc is once other factors like temp, antenna position, sat acquisition, etc, are optimized?  That's what I infer from what Skip Withrow from RDR told me about the oscillators.  He told me they can vary by a factor of 10.

Dave
N3DT

Today I moved the antenna to above the beams. This is temporary because it's in the way of the beams turning. I guess I need to get someone out here to help me install it on top of the beam mast. I don't feel confident to do it myself anymore. I suppose it's not going to hurt anything if once in a while the antenna gets rotated? Just not while I'm doing something with the Nortel. It's pretty much in the clear and as high as I can get it. It sees clear sky especially to the south except for one tree and the antenna is probably within 10' of that tree's height. There are still trees to the east that I just can't do much about. The ones to the west are about 100' away and the antenna shouldn't see them much. http://i251.photobucket.com/albums/gg287/DogTi/time/abovebeams_zpsbf8474c1.jpg So the Osc ppt is the offset from what the Nortel thinks is exactly 10MHz as I read it on page 117 of the ThunderBolt instructions (thanks Tom). How do I tell how accurate the Nortel thinks exactly 10MHz is, or can I assume it's right? If right is the case, then if I get a spread of around 300ppt, that means I'm always within 300x10^-9 Hz of 10MHz or .0003Hz at 1GHz? Or is my thinking off base and did I miss a decimal somewhere? I suppose the ppt spread is pretty much a function of how stable the osc is once other factors like temp, antenna position, sat acquisition, etc, are optimized? That's what I infer from what Skip Withrow from RDR told me about the oscillators. He told me they can vary by a factor of 10. Dave N3DT
Q
quartz55
Sat, Sep 14, 2013 5:26 PM

Here's a better view of the antenna looking due south.
http://i251.photobucket.com/albums/gg287/DogTi/abovebeamssouth_zpsd8fb78ca.jpg

Dave

Here's a better view of the antenna looking due south. http://i251.photobucket.com/albums/gg287/DogTi/abovebeamssouth_zpsd8fb78ca.jpg Dave
BC
Bob Camp
Sat, Sep 14, 2013 5:38 PM

Hi

On Sep 14, 2013, at 12:45 PM, "quartz55" quartz55@hughes.net wrote:

Today I moved the antenna to above the beams.  This is temporary because it's in the way of the beams turning.  I guess I need to get someone out here to help me install it on top of the beam mast.  I don't feel confident to do it myself anymore.  I suppose it's not going to hurt anything if once in a while the antenna gets rotated?  Just not while I'm doing something with the Nortel.  It's pretty much in the clear and as high as I can get it.  It sees clear sky especially to the south except for one tree and the antenna is probably within 10' of that tree's height.  There are still trees to the east that I just can't do much about.  The ones to the west are about 100' away and the antenna shouldn't see them much.

As long as the GPS antenna does not rotate that should be a good location. If it rotates, you will get a few inches of displacement which is less than ideal. The larger issue will be how the antenna handles lightning and the near field transmit power from the beams.

http://i251.photobucket.com/albums/gg287/DogTi/time/abovebeams_zpsbf8474c1.jpg

So the Osc ppt is the offset from what the Nortel thinks is exactly 10MHz as I read it on page 117 of the ThunderBolt instructions (thanks Tom).  How do I tell how accurate the Nortel thinks exactly 10MHz is, or can I assume it's right?  If right is the case, then if I get a spread of around 300ppt, that means I'm always within 300x10^-9 Hz of 10MHz or .0003Hz at 1GHz?

PPT = parts per trillion that's one in 1,000,000,000,000 (12 zeros). One Hz at 10 MHz is 0.1 PPM. PPM is 1 in 1,000,000 (six zeros). PPB is nine zeros, or one Hz at 1 GHz. It's also one nanosecond per second.

300 ppt would be 0.3 Hz at one GHz.

(yes, you can find a lot) of posts to the list where I've slipped a decimal on that stuff).

Or is my thinking off base and did I miss a decimal somewhere?  I suppose the ppt spread is pretty much a function of how stable the osc is once other factors like temp, antenna position, sat acquisition, etc, are optimized?

Also include the errors in the GPS signal:

  1. Uncorrected ionospheric errors
  2. GPS clock errors
  3. GPS almanac errors
  4. Antenna location errors (including interesting trivia like tides)
  5. Multi path signal errors
  6. Temperature induced errors (or not) on the antenna
  7. Drift / phase errors in the GPS receiver (there's lots of doppler on the signals).
  8. Just plain dumb decisions by the firmware

Since most of the GPS errors are going to "track out" they will not show up on the display. They will indeed shift your output frequency.

That's what I infer from what Skip Withrow from RDR told me about the oscillators.  He told me they can vary by a factor of 10.

The temperature stability of those OCXO's likely is within about 2:1. Aging wise they are like any OCXO, the longer they are kept power on, the less they will drift.

Bob

Dave
N3DT


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi On Sep 14, 2013, at 12:45 PM, "quartz55" <quartz55@hughes.net> wrote: > Today I moved the antenna to above the beams. This is temporary because it's in the way of the beams turning. I guess I need to get someone out here to help me install it on top of the beam mast. I don't feel confident to do it myself anymore. I suppose it's not going to hurt anything if once in a while the antenna gets rotated? Just not while I'm doing something with the Nortel. It's pretty much in the clear and as high as I can get it. It sees clear sky especially to the south except for one tree and the antenna is probably within 10' of that tree's height. There are still trees to the east that I just can't do much about. The ones to the west are about 100' away and the antenna shouldn't see them much. > As long as the GPS antenna does not rotate that should be a good location. If it rotates, you will get a few inches of displacement which is less than ideal. The larger issue will be how the antenna handles lightning and the near field transmit power from the beams. > http://i251.photobucket.com/albums/gg287/DogTi/time/abovebeams_zpsbf8474c1.jpg > > So the Osc ppt is the offset from what the Nortel thinks is exactly 10MHz as I read it on page 117 of the ThunderBolt instructions (thanks Tom). How do I tell how accurate the Nortel thinks exactly 10MHz is, or can I assume it's right? If right is the case, then if I get a spread of around 300ppt, that means I'm always within 300x10^-9 Hz of 10MHz or .0003Hz at 1GHz? PPT = parts per trillion that's one in 1,000,000,000,000 (12 zeros). One Hz at 10 MHz is 0.1 PPM. PPM is 1 in 1,000,000 (six zeros). PPB is nine zeros, or one Hz at 1 GHz. It's also one nanosecond per second. 300 ppt would be 0.3 Hz at one GHz. (yes, you can find a *lot*) of posts to the list where I've slipped a decimal on that stuff). > Or is my thinking off base and did I miss a decimal somewhere? I suppose the ppt spread is pretty much a function of how stable the osc is once other factors like temp, antenna position, sat acquisition, etc, are optimized? Also include the errors in the GPS signal: 1) Uncorrected ionospheric errors 2) GPS clock errors 3) GPS almanac errors 4) Antenna location errors (including interesting trivia like tides) 5) Multi path signal errors 6) Temperature induced errors (or not) on the antenna 7) Drift / phase errors in the GPS receiver (there's lots of doppler on the signals). 8) Just plain dumb decisions by the firmware Since most of the GPS errors are going to "track out" they will not show up on the display. They will indeed shift your output frequency. > That's what I infer from what Skip Withrow from RDR told me about the oscillators. He told me they can vary by a factor of 10. The temperature stability of those OCXO's likely is within about 2:1. Aging wise they are like any OCXO, the longer they are kept power on, the less they will drift. Bob > > Dave > N3DT > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BC
Bob Camp
Sat, Sep 14, 2013 5:40 PM

Hi

As long as the wind does not blow the tower around, that's probably your best location.

Bob

On Sep 14, 2013, at 1:26 PM, "quartz55" quartz55@hughes.net wrote:

Here's a better view of the antenna looking due south.
http://i251.photobucket.com/albums/gg287/DogTi/abovebeamssouth_zpsd8fb78ca.jpg

Dave


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi As long as the wind does not blow the tower around, that's probably your best location. Bob On Sep 14, 2013, at 1:26 PM, "quartz55" <quartz55@hughes.net> wrote: > Here's a better view of the antenna looking due south. > http://i251.photobucket.com/albums/gg287/DogTi/abovebeamssouth_zpsd8fb78ca.jpg > > Dave > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
TV
Tom Van Baak
Sat, Sep 14, 2013 5:51 PM

I get a spread of around 300ppt, that means I'm always within
300x10^-9 Hz of 10MHz or .0003Hz at 1GHz?

Note "300 ppt" is a "fractional frequency", unit-less value, so
at 10 MHz, 300e-12 * 1e7 Hz =  0.003 Hz
at 1 GHz, 300e-12 * 1e9 Hz = 0.3 Hz

I suppose the ppt spread is pretty much a function of how stable the osc is once
other factors like temp, antenna position, sat acquisition, etc, are optimized?

Yes and no; it's a complicated subject. For now, just two points:

A measure of frequency implies some measurement duration. A given TBolt may be off by more than 1e-11 in frequency over seconds or minutes even though if you measure over a day it is accurate to less than 1e-13. This is one problem with interpreting the "OSC" value.

A second problem is that a TBolt can't really know its own accuracy; that requires an external frequency reference, so take the PPS and OSC with a grain of salt. However, in general, the TBolt will steer its 10 MHz oscillator so as to minimize the PPS and OSC values. As such they can be used as a rough idea of how well the unit is performing.

/tvb

> I get a spread of around 300ppt, that means I'm always within > 300x10^-9 Hz of 10MHz or .0003Hz at 1GHz? Note "300 ppt" is a "fractional frequency", unit-less value, so at 10 MHz, 300e-12 * 1e7 Hz = 0.003 Hz at 1 GHz, 300e-12 * 1e9 Hz = 0.3 Hz > I suppose the ppt spread is pretty much a function of how stable the osc is once > other factors like temp, antenna position, sat acquisition, etc, are optimized? Yes and no; it's a complicated subject. For now, just two points: A measure of frequency implies some measurement duration. A given TBolt may be off by more than 1e-11 in frequency over seconds or minutes even though if you measure over a day it is accurate to less than 1e-13. This is one problem with interpreting the "OSC" value. A second problem is that a TBolt can't really know its own accuracy; that requires an external frequency reference, so take the PPS and OSC with a grain of salt. However, in general, the TBolt will steer its 10 MHz oscillator so as to minimize the PPS and OSC values. As such they can be used as a rough idea of how well the unit is performing. /tvb
Q
quartz55
Sat, Sep 14, 2013 6:59 PM

If it's on the top of the antenna mast, it should only rotate around the mast, which may be 4" at most. I can stabalize the mast a bit with a outrigger at the bottom which I've been meaning to do for years.  The pipe from the chimney is 1" thick wall, so there isn't much wind effect there, I forget what the mast is, whatever fit in the Alliance rotator.  It survived Sandy which brought down a huge oak on the roof and whacked the very end director of the 2M beam and bent one side down.  If you're interested http://s251.photobucket.com/user/DogTi/library/Tree?sort=4&page=1

Thanks, yeah I always get messed up with decimals and powers of 10, especially when they flip from + to -, unless I work with them all the time, plus I've got old age brain rot.  It ain't what it used to be.

I'm doing a precision survey now.

Dave

If it's on the top of the antenna mast, it should only rotate around the mast, which may be 4" at most. I can stabalize the mast a bit with a outrigger at the bottom which I've been meaning to do for years. The pipe from the chimney is 1" thick wall, so there isn't much wind effect there, I forget what the mast is, whatever fit in the Alliance rotator. It survived Sandy which brought down a huge oak on the roof and whacked the very end director of the 2M beam and bent one side down. If you're interested http://s251.photobucket.com/user/DogTi/library/Tree?sort=4&page=1 Thanks, yeah I always get messed up with decimals and powers of 10, especially when they flip from + to -, unless I work with them all the time, plus I've got old age brain rot. It ain't what it used to be. I'm doing a precision survey now. Dave
BC
Bob Camp
Sat, Sep 14, 2013 7:20 PM

Hi

If you mostly keep the beams pointed into the wind, I'd do that before you do the next survey.

Bob

On Sep 14, 2013, at 2:59 PM, "quartz55" quartz55@hughes.net wrote:

If it's on the top of the antenna mast, it should only rotate around the mast, which may be 4" at most. I can stabalize the mast a bit with a outrigger at the bottom which I've been meaning to do for years.  The pipe from the chimney is 1" thick wall, so there isn't much wind effect there, I forget what the mast is, whatever fit in the Alliance rotator.  It survived Sandy which brought down a huge oak on the roof and whacked the very end director of the 2M beam and bent one side down.  If you're interested http://s251.photobucket.com/user/DogTi/library/Tree?sort=4&page=1

Thanks, yeah I always get messed up with decimals and powers of 10, especially when they flip from + to -, unless I work with them all the time, plus I've got old age brain rot.  It ain't what it used to be.

I'm doing a precision survey now.

Dave


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi If you mostly keep the beams pointed into the wind, I'd do that before you do the next survey. Bob On Sep 14, 2013, at 2:59 PM, "quartz55" <quartz55@hughes.net> wrote: > If it's on the top of the antenna mast, it should only rotate around the mast, which may be 4" at most. I can stabalize the mast a bit with a outrigger at the bottom which I've been meaning to do for years. The pipe from the chimney is 1" thick wall, so there isn't much wind effect there, I forget what the mast is, whatever fit in the Alliance rotator. It survived Sandy which brought down a huge oak on the roof and whacked the very end director of the 2M beam and bent one side down. If you're interested http://s251.photobucket.com/user/DogTi/library/Tree?sort=4&page=1 > > Thanks, yeah I always get messed up with decimals and powers of 10, especially when they flip from + to -, unless I work with them all the time, plus I've got old age brain rot. It ain't what it used to be. > > I'm doing a precision survey now. > > Dave > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
W
WarrenS
Sat, Sep 14, 2013 10:21 PM

OT for the current heading, so I renamed it "Oscillator stability"

as Tom says, "it's a complicated subject".
And to complicate things even further, here are a few of advanced subtleties
that I've observed from the TBolt when using LadyHeather.

  1. The TBolt uses the received GPS signal as the reference for the
    calculated OSC freq offset and the Phase offset, over a measurement time of
    1 sec.
    Unfortunately, from second to second the GPS signal is very noisy.
    Fortunately, LadyHeather can display plots of the data with several helpful
    user options, such as gain, offset, and most important length of time
    averaged.

  2. The Osc freq display plot of LH is so noisy that I find it mostly
    useless,
    unless I turn on LH's display filter to average the data over more than 1
    second.
    LH allows you to manually turn-on and set the display filter to any time
    period desired,
    doing so this will greatly reduce the noise of the Osc plot.
    (10 samples is the display filter default, but I find 100sec to be a better
    place to start, for most things)
    A problem with the LH Osc freq plot even after filtering, is that the
    frequency difference data can not be counted on to be more accurate than a
    few parts in 1e-12 due to some offset rounding problems that often occurs.

  3. On the other hand, the filtered Phase plot has no known offset error.
    The Tbolt accurately shows time/phase different between the 1PPS and the
    received GPS signal,
    and when disciplined, it assumes if there is a difference, then the GPS is
    always right.
    This is why the antenna placement and setup is so important. Gunk in, Gunk
    out.

  4. The GPS signal, even on a 'perfect antenna', tends to wonders around ~10
    ns PP independently of the time period averaged.
    So if LH is showing less than ~ 10 ns of GPS noise in the phase plot,
    it is because the control loop is set too fast and therefore forcing the
    Osc's freq to change a little so that it's phase will wonder around with the
    GPS's noise.
    Tbolt's max useable time constant is only 1000 sec, which is not nearly long
    enough to avoid this problem, when using a stable external osc like a good
    RB.
    To avoid tracking the noisy GPS data, the extended TC method must be used to
    set the Tbolt's osc to values >> 1000 seconds,
    and/or you can reduce the speed of  the phase tracking using the damping
    setting..

  5. How well and how fast the Tbolt minimizes the PPS phase and freq offset
    compared to the received GPS signal all depends on tuning.
    The Tbolt's TC determines what the Frequency  tracking time constant will
    be, and the TBolt's damping setting determines what the Phase tracking time
    multiplier is.
    If you set the damping factor to a large value like 100, then the Phase
    tracking will pretty much be turned off,
    making the disciplined loop more of a freq lock loop instead of a phase lock
    loop. This is done by lowering the gain of the loop's PID integrator.
    This is a way to set the phase's tracking time constant to be much slower
    than the freq TC setting, and if desired the phase tracking TC can be made
    several days long.
    Turning off the phase tracking has it's own set of pros and cons, but in
    most cases it is generally not desirerable in GPSDO.
    A damping setting of 0.7 to 1 will give the best overall compromise between
    the trade-off of not adding extra freq noise but still allowing good phase
    tracking.

  6. What some do not realize is to correct for any phase drift error, the
    Oscillator must be set off frequency.
    When the frequency is correct there is no further change in the present
    phase, whatever the present phase may be or wherever it may of come from.
    The faster you want to correct or change the present phase difference, no
    mater how it got there, the larger that the present frequency error must be
    made. (this causes freq noise)
    The trade off is, if you do not correct the present phase error then the
    past average frequency will be in error.

  7. So it is all a matter of what is more important to the application,
    present frequency error and noise or the average of all past frequency
    errors?
    The goal of most GPSDO is to keep the average past frequency error to zero.
    (a Phase Lock Loop)
    Where as for many transmitter things such as used by Hams,  it is the
    present errors and noise that is more important.
    So no need to cause a present freq error just to fix something that happened
    in the past.
    The past is gone and what happened before does not matter anymore. (a Freq
    Lock Loop)

ws


----- Original Message -----
From: "Tom Van Baak"
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Subject: Re: [time-nuts] New NTBW50AA

I get a spread of around 300ppt, that means I'm always within
300x10^-9 Hz of 10MHz or .0003Hz at 1GHz?

Note "300 ppt" is a "fractional frequency", unit-less value, so
at 10 MHz, 300e-12 * 1e7 Hz =  0.003 Hz
at 1 GHz, 300e-12 * 1e9 Hz = 0.3 Hz

I suppose the ppt spread is pretty much a function of how stable the osc
is once
other factors like temp, antenna position, sat acquisition, etc, are
optimized?

Yes and no; it's a complicated subject. For now, just two points:

A measure of frequency implies some measurement duration. A given TBolt may
be off by more than 1e-11 in frequency over seconds or minutes even though
if you measure over a day it is accurate to less than 1e-13. This is one
problem with interpreting the "OSC" value.

A second problem is that a TBolt can't really know its own accuracy; that
requires an external frequency reference, so take the PPS and OSC with a
grain of salt. However, in general, the TBolt will steer its 10 MHz
oscillator so as to minimize the PPS and OSC values. As such they can be
used as a rough idea of how well the unit is performing.

/tvb

OT for the current heading, so I renamed it "Oscillator stability" as Tom says, "it's a complicated subject". And to complicate things even further, here are a few of advanced subtleties that I've observed from the TBolt when using LadyHeather. 1) The TBolt uses the received GPS signal as the reference for the calculated OSC freq offset and the Phase offset, over a measurement time of 1 sec. Unfortunately, from second to second the GPS signal is very noisy. Fortunately, LadyHeather can display plots of the data with several helpful user options, such as gain, offset, and most important length of time averaged. 2) The Osc freq display plot of LH is so noisy that I find it mostly useless, unless I turn on LH's display filter to average the data over more than 1 second. LH allows you to manually turn-on and set the display filter to any time period desired, doing so this will greatly reduce the noise of the Osc plot. (10 samples is the display filter default, but I find 100sec to be a better place to start, for most things) A problem with the LH Osc freq plot even after filtering, is that the frequency difference data can not be counted on to be more accurate than a few parts in 1e-12 due to some offset rounding problems that often occurs. 3) On the other hand, the filtered Phase plot has no known offset error. The Tbolt accurately shows time/phase different between the 1PPS and the received GPS signal, and when disciplined, it assumes if there is a difference, then the GPS is always right. This is why the antenna placement and setup is so important. Gunk in, Gunk out. 4) The GPS signal, even on a 'perfect antenna', tends to wonders around ~10 ns PP independently of the time period averaged. So if LH is showing less than ~ 10 ns of GPS noise in the phase plot, it is because the control loop is set too fast and therefore forcing the Osc's freq to change a little so that it's phase will wonder around with the GPS's noise. Tbolt's max useable time constant is only 1000 sec, which is not nearly long enough to avoid this problem, when using a stable external osc like a good RB. To avoid tracking the noisy GPS data, the extended TC method must be used to set the Tbolt's osc to values >> 1000 seconds, and/or you can reduce the speed of the phase tracking using the damping setting.. 4) How well and how fast the Tbolt minimizes the PPS phase and freq offset compared to the received GPS signal all depends on tuning. The Tbolt's TC determines what the Frequency tracking time constant will be, and the TBolt's damping setting determines what the Phase tracking time multiplier is. If you set the damping factor to a large value like 100, then the Phase tracking will pretty much be turned off, making the disciplined loop more of a freq lock loop instead of a phase lock loop. This is done by lowering the gain of the loop's PID integrator. This is a way to set the phase's tracking time constant to be much slower than the freq TC setting, and if desired the phase tracking TC can be made several days long. Turning off the phase tracking has it's own set of pros and cons, but in most cases it is generally not desirerable in GPSDO. A damping setting of 0.7 to 1 will give the best overall compromise between the trade-off of not adding extra freq noise but still allowing good phase tracking. 5) What some do not realize is to correct for any phase drift error, the Oscillator must be set off frequency. When the frequency is correct there is no further change in the present phase, whatever the present phase may be or wherever it may of come from. The faster you want to correct or change the present phase difference, no mater how it got there, the larger that the present frequency error must be made. (this causes freq noise) The trade off is, if you do not correct the present phase error then the past average frequency will be in error. 6) So it is all a matter of what is more important to the application, present frequency error and noise or the average of all past frequency errors? The goal of most GPSDO is to keep the average past frequency error to zero. (a Phase Lock Loop) Where as for many transmitter things such as used by Hams, it is the present errors and noise that is more important. So no need to cause a present freq error just to fix something that happened in the past. The past is gone and what happened before does not matter anymore. (a Freq Lock Loop) ws *********************** ----- Original Message ----- From: "Tom Van Baak" To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com> Subject: Re: [time-nuts] New NTBW50AA > I get a spread of around 300ppt, that means I'm always within > 300x10^-9 Hz of 10MHz or .0003Hz at 1GHz? Note "300 ppt" is a "fractional frequency", unit-less value, so at 10 MHz, 300e-12 * 1e7 Hz = 0.003 Hz at 1 GHz, 300e-12 * 1e9 Hz = 0.3 Hz > I suppose the ppt spread is pretty much a function of how stable the osc > is once > other factors like temp, antenna position, sat acquisition, etc, are > optimized? Yes and no; it's a complicated subject. For now, just two points: A measure of frequency implies some measurement duration. A given TBolt may be off by more than 1e-11 in frequency over seconds or minutes even though if you measure over a day it is accurate to less than 1e-13. This is one problem with interpreting the "OSC" value. A second problem is that a TBolt can't really know its own accuracy; that requires an external frequency reference, so take the PPS and OSC with a grain of salt. However, in general, the TBolt will steer its 10 MHz oscillator so as to minimize the PPS and OSC values. As such they can be used as a rough idea of how well the unit is performing. /tvb
BS
Bob Stewart
Sat, Sep 14, 2013 10:59 PM

Hi Warren,

There's a lot of good information in this post, and I've discovered a lot of it while writing my GPSDO PLL.  In point 5, you mention that to correct for frequency it has to be off frequency.  I've seen this on both sides of it.  From time to time, I see extremely long periods of no DAC activity.  And at other times, it's moving back and forth; apparently because the DAC is about 1 bit too small.  In my PLL, I've just decided to accept the phase oscillation when it happens and blame it on two main factors.  One is the size in bits of the DAC, about which there's little I can do.  (Well, Jacques Audet came up with something to get one more bit out of it.) 

The other is the Oncore sawtooth, which I'll start working on sometime this week.  The idea of keeping the phase a bit off from the PPS is just a bit too much to bother with for me, I think.  All I have to work with is an 18F2220 PIC running at 40MHz, and I'd like to keep it as simple as possible.  It's already a lot more accurate than I had expected to be able to get.  And it's not like the slope detector on this thing is all that reliable down in the single digit nanoseconds.

Sorry for the thread steal.  Thanks for the review.

Bob - AE6RV


From: WarrenS warrensjmail-one@yahoo.com
To: Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Saturday, September 14, 2013 5:21 PM
Subject: Re: [time-nuts] Oscillator stability  was (New NTBW50AA)

OT for the current heading, so I renamed it "Oscillator stability"

as Tom says, "it's a complicated subject".
And to complicate things even further, here are a few of advanced subtleties that I've observed from the TBolt when using LadyHeather.

  1. The TBolt uses the received GPS signal as the reference for the calculated OSC freq offset and the Phase offset, over a measurement time of 1 sec.
    Unfortunately, from second to second the GPS signal is very noisy.
    Fortunately, LadyHeather can display plots of the data with several helpful user options, such as gain, offset, and most important length of time averaged.

  2. The Osc freq display plot of LH is so noisy that I find it mostly useless,
    unless I turn on LH's display filter to average the data over more than 1 second.
    LH allows you to manually turn-on and set the display filter to any time period desired,
    doing so this will greatly reduce the noise of the Osc plot.
    (10 samples is the display filter default, but I find 100sec to be a better place to start, for most things)
    A problem with the LH Osc freq plot even after filtering, is that the frequency difference data can not be counted on to be more accurate than a few parts in 1e-12 due to some offset rounding problems that often occurs.

  3. On the other hand, the filtered Phase plot has no known offset error.
    The Tbolt accurately shows time/phase different between the 1PPS and the received GPS signal,
    and when disciplined, it assumes if there is a difference, then the GPS is always right.
    This is why the antenna placement and setup is so important. Gunk in, Gunk out.

  4. The GPS signal, even on a 'perfect antenna', tends to wonders around ~10 ns PP independently of the time period averaged.
    So if LH is showing less than ~ 10 ns of GPS noise in the phase plot,
    it is because the control loop is set too fast and therefore forcing the Osc's freq to change a little so that it's phase will wonder around with the GPS's noise.
    Tbolt's max useable time constant is only 1000 sec, which is not nearly long enough to avoid this problem, when using a stable external osc like a good RB.
    To avoid tracking the noisy GPS data, the extended TC method must be used to set the Tbolt's osc to values >> 1000 seconds,
    and/or you can reduce the speed of  the phase tracking using the damping setting..

  5. How well and how fast the Tbolt minimizes the PPS phase and freq offset compared to the received GPS signal all depends on tuning.
    The Tbolt's TC determines what the Frequency  tracking time constant will be, and the TBolt's damping setting determines what the Phase tracking time multiplier is.
    If you set the damping factor to a large value like 100, then the Phase tracking will pretty much be turned off,
    making the disciplined loop more of a freq lock loop instead of a phase lock loop. This is done by lowering the gain of the loop's PID integrator.
    This is a way to set the phase's tracking time constant to be much slower than the freq TC setting, and if desired the phase tracking TC can be made several days long.
    Turning off the phase tracking has it's own set of pros and cons, but in most cases it is generally not desirerable in GPSDO.
    A damping setting of 0.7 to 1 will give the best overall compromise between the trade-off of not adding extra freq noise but still allowing good phase tracking.

  6. What some do not realize is to correct for any phase drift error, the Oscillator must be set off frequency.
    When the frequency is correct there is no further change in the present phase, whatever the present phase may be or wherever it may of come from.
    The faster you want to correct or change the present phase difference, no mater how it got there, the larger that the present frequency error must be made. (this causes freq noise)
    The trade off is, if you do not correct the present phase error then the past average frequency will be in error.

  7. So it is all a matter of what is more important to the application, present frequency error and noise or the average of all past frequency errors?
    The goal of most GPSDO is to keep the average past frequency error to zero. (a Phase Lock Loop)
    Where as for many transmitter things such as used by Hams,  it is the present errors and noise that is more important.
    So no need to cause a present freq error just to fix something that happened in the past.
    The past is gone and what happened before does not matter anymore. (a Freq Lock Loop)

ws


----- Original Message ----- From: "Tom Van Baak"
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Subject: Re: [time-nuts] New NTBW50AA

I get a spread of around 300ppt, that means I'm always within
300x10^-9 Hz of 10MHz or .0003Hz at 1GHz?

Note "300 ppt" is a "fractional frequency", unit-less value, so
at 10 MHz, 300e-12 * 1e7 Hz =  0.003 Hz
at 1 GHz, 300e-12 * 1e9 Hz = 0.3 Hz

I suppose the ppt spread is pretty much a function of how stable the osc is once
other factors like temp, antenna position, sat acquisition, etc, are optimized?

Yes and no; it's a complicated subject. For now, just two points:

A measure of frequency implies some measurement duration. A given TBolt may be off by more than 1e-11 in frequency over seconds or minutes even though if you measure over a day it is accurate to less than 1e-13. This is one problem with interpreting the "OSC" value.

A second problem is that a TBolt can't really know its own accuracy; that requires an external frequency reference, so take the PPS and OSC with a grain of salt. However, in general, the TBolt will steer its 10 MHz oscillator so as to minimize the PPS and OSC values. As such they can be used as a rough idea of how well the unit is performing.

/tvb


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Warren, There's a lot of good information in this post, and I've discovered a lot of it while writing my GPSDO PLL.  In point 5, you mention that to correct for frequency it has to be off frequency.  I've seen this on both sides of it.  From time to time, I see extremely long periods of no DAC activity.  And at other times, it's moving back and forth; apparently because the DAC is about 1 bit too small.  In my PLL, I've just decided to accept the phase oscillation when it happens and blame it on two main factors.  One is the size in bits of the DAC, about which there's little I can do.  (Well, Jacques Audet came up with something to get one more bit out of it.)  The other is the Oncore sawtooth, which I'll start working on sometime this week.  The idea of keeping the phase a bit off from the PPS is just a bit too much to bother with for me, I think.  All I have to work with is an 18F2220 PIC running at 40MHz, and I'd like to keep it as simple as possible.  It's already a lot more accurate than I had expected to be able to get.  And it's not like the slope detector on this thing is all that reliable down in the single digit nanoseconds. Sorry for the thread steal.  Thanks for the review. Bob - AE6RV >________________________________ > From: WarrenS <warrensjmail-one@yahoo.com> >To: Discussion of precise time and frequency measurement <time-nuts@febo.com> >Sent: Saturday, September 14, 2013 5:21 PM >Subject: Re: [time-nuts] Oscillator stability was (New NTBW50AA) > > >OT for the current heading, so I renamed it "Oscillator stability" > >as Tom says, "it's a complicated subject". >And to complicate things even further, here are a few of advanced subtleties that I've observed from the TBolt when using LadyHeather. > > >1) The TBolt uses the received GPS signal as the reference for the calculated OSC freq offset and the Phase offset, over a measurement time of 1 sec. >Unfortunately, from second to second the GPS signal is very noisy. >Fortunately, LadyHeather can display plots of the data with several helpful user options, such as gain, offset, and most important length of time averaged. > >2) The Osc freq display plot of LH is so noisy that I find it mostly useless, >unless I turn on LH's display filter to average the data over more than 1 second. >LH allows you to manually turn-on and set the display filter to any time period desired, >doing so this will greatly reduce the noise of the Osc plot. >(10 samples is the display filter default, but I find 100sec to be a better place to start, for most things) >A problem with the LH Osc freq plot even after filtering, is that the frequency difference data can not be counted on to be more accurate than a few parts in 1e-12 due to some offset rounding problems that often occurs. > > >3) On the other hand, the filtered Phase plot has no known offset error. >The Tbolt accurately shows time/phase different between the 1PPS and the received GPS signal, >and when disciplined, it assumes if there is a difference, then the GPS is always right. >This is why the antenna placement and setup is so important. Gunk in, Gunk out. > > >4) The GPS signal, even on a 'perfect antenna', tends to wonders around ~10 ns PP independently of the time period averaged. >So if LH is showing less than ~ 10 ns of GPS noise in the phase plot, >it is because the control loop is set too fast and therefore forcing the Osc's freq to change a little so that it's phase will wonder around with the GPS's noise. >Tbolt's max useable time constant is only 1000 sec, which is not nearly long enough to avoid this problem, when using a stable external osc like a good RB. >To avoid tracking the noisy GPS data, the extended TC method must be used to set the Tbolt's osc to values >> 1000 seconds, >and/or you can reduce the speed of  the phase tracking using the damping setting.. > >4) How well and how fast the Tbolt minimizes the PPS phase and freq offset compared to the received GPS signal all depends on tuning. >The Tbolt's TC determines what the Frequency  tracking time constant will be, and the TBolt's damping setting determines what the Phase tracking time multiplier is. >If you set the damping factor to a large value like 100, then the Phase tracking will pretty much be turned off, >making the disciplined loop more of a freq lock loop instead of a phase lock loop. This is done by lowering the gain of the loop's PID integrator. >This is a way to set the phase's tracking time constant to be much slower than the freq TC setting, and if desired the phase tracking TC can be made several days long. >Turning off the phase tracking has it's own set of pros and cons, but in most cases it is generally not desirerable in GPSDO. >A damping setting of 0.7 to 1 will give the best overall compromise between the trade-off of not adding extra freq noise but still allowing good phase tracking. > > >5) What some do not realize is to correct for any phase drift error, the Oscillator must be set off frequency. >When the frequency is correct there is no further change in the present phase, whatever the present phase may be or wherever it may of come from. >The faster you want to correct or change the present phase difference, no mater how it got there, the larger that the present frequency error must be made. (this causes freq noise) >The trade off is, if you do not correct the present phase error then the past average frequency will be in error. > >6) So it is all a matter of what is more important to the application, present frequency error and noise or the average of all past frequency errors? >The goal of most GPSDO is to keep the average past frequency error to zero. (a Phase Lock Loop) >Where as for many transmitter things such as used by Hams,  it is the present errors and noise that is more important. >So no need to cause a present freq error just to fix something that happened in the past. >The past is gone and what happened before does not matter anymore. (a Freq Lock Loop) > >ws > >*********************** > >----- Original Message ----- From: "Tom Van Baak" >To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com> >Subject: Re: [time-nuts] New NTBW50AA > > >> I get a spread of around 300ppt, that means I'm always within >> 300x10^-9 Hz of 10MHz or .0003Hz at 1GHz? > >Note "300 ppt" is a "fractional frequency", unit-less value, so >at 10 MHz, 300e-12 * 1e7 Hz =  0.003 Hz >at 1 GHz, 300e-12 * 1e9 Hz = 0.3 Hz > >> I suppose the ppt spread is pretty much a function of how stable the osc is once >> other factors like temp, antenna position, sat acquisition, etc, are optimized? > >Yes and no; it's a complicated subject. For now, just two points: > >A measure of frequency implies some measurement duration. A given TBolt may be off by more than 1e-11 in frequency over seconds or minutes even though if you measure over a day it is accurate to less than 1e-13. This is one problem with interpreting the "OSC" value. > >A second problem is that a TBolt can't really know its own accuracy; that requires an external frequency reference, so take the PPS and OSC with a grain of salt. However, in general, the TBolt will steer its 10 MHz oscillator so as to minimize the PPS and OSC values. As such they can be used as a rough idea of how well the unit is performing. > >/tvb > > > >_______________________________________________ >time-nuts mailing list -- time-nuts@febo.com >To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >and follow the instructions there. > > >
RK
Rob Kimberley
Sun, Sep 15, 2013 10:09 AM

Do you have any idea how much the antenna will move in a strong wind? It
really needs to be in a secure rigid position for optimum timing
performance. Sorry if this has been covered before but just picked up the
thread.

Rob

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of quartz55
Sent: 14 September 2013 18:26
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] New NTBW50AA

Here's a better view of the antenna looking due south.
http://i251.photobucket.com/albums/gg287/DogTi/abovebeamssouth_zpsd8fb78ca.j
pg

Dave


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Do you have any idea how much the antenna will move in a strong wind? It really needs to be in a secure rigid position for optimum timing performance. Sorry if this has been covered before but just picked up the thread. Rob -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On Behalf Of quartz55 Sent: 14 September 2013 18:26 To: Discussion of precise time and frequency measurement Subject: Re: [time-nuts] New NTBW50AA Here's a better view of the antenna looking due south. http://i251.photobucket.com/albums/gg287/DogTi/abovebeamssouth_zpsd8fb78ca.j pg Dave _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
JS
Jim Sanford
Sun, Sep 15, 2013 11:14 AM

So, what settings would you recommend for an application where frequency
accuracy and low phase noise of the 10 MHz output are the objectives?

In particular, I'm not sure I understand how to establish the "extended
TC method."

Thanks!
Jim
wb4gcs@amsat.org

On 9/14/2013 6:21 PM, WarrenS wrote:

OT for the current heading, so I renamed it "Oscillator stability"

as Tom says, "it's a complicated subject".
And to complicate things even further, here are a few of advanced
subtleties that I've observed from the TBolt when using LadyHeather.

  1. The TBolt uses the received GPS signal as the reference for the
    calculated OSC freq offset and the Phase offset, over a measurement
    time of 1 sec.
    Unfortunately, from second to second the GPS signal is very noisy.
    Fortunately, LadyHeather can display plots of the data with several
    helpful user options, such as gain, offset, and most important length
    of time averaged.

  2. The Osc freq display plot of LH is so noisy that I find it mostly
    useless,
    unless I turn on LH's display filter to average the data over more
    than 1 second.
    LH allows you to manually turn-on and set the display filter to any
    time period desired,
    doing so this will greatly reduce the noise of the Osc plot.
    (10 samples is the display filter default, but I find 100sec to be a
    better place to start, for most things)
    A problem with the LH Osc freq plot even after filtering, is that the
    frequency difference data can not be counted on to be more accurate
    than a few parts in 1e-12 due to some offset rounding problems that
    often occurs.

  3. On the other hand, the filtered Phase plot has no known offset error.
    The Tbolt accurately shows time/phase different between the 1PPS and
    the received GPS signal,
    and when disciplined, it assumes if there is a difference, then the
    GPS is always right.
    This is why the antenna placement and setup is so important. Gunk in,
    Gunk out.

  4. The GPS signal, even on a 'perfect antenna', tends to wonders
    around ~10 ns PP independently of the time period averaged.
    So if LH is showing less than ~ 10 ns of GPS noise in the phase plot,
    it is because the control loop is set too fast and therefore forcing
    the Osc's freq to change a little so that it's phase will wonder
    around with the GPS's noise.
    Tbolt's max useable time constant is only 1000 sec, which is not
    nearly long enough to avoid this problem, when using a stable external
    osc like a good RB.
    To avoid tracking the noisy GPS data, the extended TC method must be
    used to set the Tbolt's osc to values >> 1000 seconds,
    and/or you can reduce the speed of  the phase tracking using the
    damping setting..

  5. How well and how fast the Tbolt minimizes the PPS phase and freq
    offset compared to the received GPS signal all depends on tuning.
    The Tbolt's TC determines what the Frequency  tracking time constant
    will be, and the TBolt's damping setting determines what the Phase
    tracking time multiplier is.
    If you set the damping factor to a large value like 100, then the
    Phase tracking will pretty much be turned off,
    making the disciplined loop more of a freq lock loop instead of a
    phase lock loop. This is done by lowering the gain of the loop's PID
    integrator.
    This is a way to set the phase's tracking time constant to be much
    slower than the freq TC setting, and if desired the phase tracking TC
    can be made several days long.
    Turning off the phase tracking has it's own set of pros and cons, but
    in most cases it is generally not desirerable in GPSDO.
    A damping setting of 0.7 to 1 will give the best overall compromise
    between the trade-off of not adding extra freq noise but still
    allowing good phase tracking.

  6. What some do not realize is to correct for any phase drift error,
    the Oscillator must be set off frequency.
    When the frequency is correct there is no further change in the
    present phase, whatever the present phase may be or wherever it may of
    come from.
    The faster you want to correct or change the present phase difference,
    no mater how it got there, the larger that the present frequency error
    must be made. (this causes freq noise)
    The trade off is, if you do not correct the present phase error then
    the past average frequency will be in error.

  7. So it is all a matter of what is more important to the application,
    present frequency error and noise or the average of all past frequency
    errors?
    The goal of most GPSDO is to keep the average past frequency error to
    zero. (a Phase Lock Loop)
    Where as for many transmitter things such as used by Hams,  it is the
    present errors and noise that is more important.
    So no need to cause a present freq error just to fix something that
    happened in the past.
    The past is gone and what happened before does not matter anymore. (a
    Freq Lock Loop)

ws


----- Original Message ----- From: "Tom Van Baak"
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Subject: Re: [time-nuts] New NTBW50AA

I get a spread of around 300ppt, that means I'm always within
300x10^-9 Hz of 10MHz or .0003Hz at 1GHz?

Note "300 ppt" is a "fractional frequency", unit-less value, so
at 10 MHz, 300e-12 * 1e7 Hz =  0.003 Hz
at 1 GHz, 300e-12 * 1e9 Hz = 0.3 Hz

I suppose the ppt spread is pretty much a function of how stable the
osc is once
other factors like temp, antenna position, sat acquisition, etc, are
optimized?

Yes and no; it's a complicated subject. For now, just two points:

A measure of frequency implies some measurement duration. A given
TBolt may be off by more than 1e-11 in frequency over seconds or
minutes even though if you measure over a day it is accurate to less
than 1e-13. This is one problem with interpreting the "OSC" value.

A second problem is that a TBolt can't really know its own accuracy;
that requires an external frequency reference, so take the PPS and OSC
with a grain of salt. However, in general, the TBolt will steer its 10
MHz oscillator so as to minimize the PPS and OSC values. As such they
can be used as a rough idea of how well the unit is performing.

/tvb


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

So, what settings would you recommend for an application where frequency accuracy and low phase noise of the 10 MHz output are the objectives? In particular, I'm not sure I understand how to establish the "extended TC method." Thanks! Jim wb4gcs@amsat.org On 9/14/2013 6:21 PM, WarrenS wrote: > OT for the current heading, so I renamed it "Oscillator stability" > > as Tom says, "it's a complicated subject". > And to complicate things even further, here are a few of advanced > subtleties that I've observed from the TBolt when using LadyHeather. > > > 1) The TBolt uses the received GPS signal as the reference for the > calculated OSC freq offset and the Phase offset, over a measurement > time of 1 sec. > Unfortunately, from second to second the GPS signal is very noisy. > Fortunately, LadyHeather can display plots of the data with several > helpful user options, such as gain, offset, and most important length > of time averaged. > > 2) The Osc freq display plot of LH is so noisy that I find it mostly > useless, > unless I turn on LH's display filter to average the data over more > than 1 second. > LH allows you to manually turn-on and set the display filter to any > time period desired, > doing so this will greatly reduce the noise of the Osc plot. > (10 samples is the display filter default, but I find 100sec to be a > better place to start, for most things) > A problem with the LH Osc freq plot even after filtering, is that the > frequency difference data can not be counted on to be more accurate > than a few parts in 1e-12 due to some offset rounding problems that > often occurs. > > > 3) On the other hand, the filtered Phase plot has no known offset error. > The Tbolt accurately shows time/phase different between the 1PPS and > the received GPS signal, > and when disciplined, it assumes if there is a difference, then the > GPS is always right. > This is why the antenna placement and setup is so important. Gunk in, > Gunk out. > > > 4) The GPS signal, even on a 'perfect antenna', tends to wonders > around ~10 ns PP independently of the time period averaged. > So if LH is showing less than ~ 10 ns of GPS noise in the phase plot, > it is because the control loop is set too fast and therefore forcing > the Osc's freq to change a little so that it's phase will wonder > around with the GPS's noise. > Tbolt's max useable time constant is only 1000 sec, which is not > nearly long enough to avoid this problem, when using a stable external > osc like a good RB. > To avoid tracking the noisy GPS data, the extended TC method must be > used to set the Tbolt's osc to values >> 1000 seconds, > and/or you can reduce the speed of the phase tracking using the > damping setting.. > > 4) How well and how fast the Tbolt minimizes the PPS phase and freq > offset compared to the received GPS signal all depends on tuning. > The Tbolt's TC determines what the Frequency tracking time constant > will be, and the TBolt's damping setting determines what the Phase > tracking time multiplier is. > If you set the damping factor to a large value like 100, then the > Phase tracking will pretty much be turned off, > making the disciplined loop more of a freq lock loop instead of a > phase lock loop. This is done by lowering the gain of the loop's PID > integrator. > This is a way to set the phase's tracking time constant to be much > slower than the freq TC setting, and if desired the phase tracking TC > can be made several days long. > Turning off the phase tracking has it's own set of pros and cons, but > in most cases it is generally not desirerable in GPSDO. > A damping setting of 0.7 to 1 will give the best overall compromise > between the trade-off of not adding extra freq noise but still > allowing good phase tracking. > > > 5) What some do not realize is to correct for any phase drift error, > the Oscillator must be set off frequency. > When the frequency is correct there is no further change in the > present phase, whatever the present phase may be or wherever it may of > come from. > The faster you want to correct or change the present phase difference, > no mater how it got there, the larger that the present frequency error > must be made. (this causes freq noise) > The trade off is, if you do not correct the present phase error then > the past average frequency will be in error. > > 6) So it is all a matter of what is more important to the application, > present frequency error and noise or the average of all past frequency > errors? > The goal of most GPSDO is to keep the average past frequency error to > zero. (a Phase Lock Loop) > Where as for many transmitter things such as used by Hams, it is the > present errors and noise that is more important. > So no need to cause a present freq error just to fix something that > happened in the past. > The past is gone and what happened before does not matter anymore. (a > Freq Lock Loop) > > ws > > *********************** > > ----- Original Message ----- From: "Tom Van Baak" > To: "Discussion of precise time and frequency measurement" > <time-nuts@febo.com> > Subject: Re: [time-nuts] New NTBW50AA > > >> I get a spread of around 300ppt, that means I'm always within >> 300x10^-9 Hz of 10MHz or .0003Hz at 1GHz? > > Note "300 ppt" is a "fractional frequency", unit-less value, so > at 10 MHz, 300e-12 * 1e7 Hz = 0.003 Hz > at 1 GHz, 300e-12 * 1e9 Hz = 0.3 Hz > >> I suppose the ppt spread is pretty much a function of how stable the >> osc is once >> other factors like temp, antenna position, sat acquisition, etc, are >> optimized? > > Yes and no; it's a complicated subject. For now, just two points: > > A measure of frequency implies some measurement duration. A given > TBolt may be off by more than 1e-11 in frequency over seconds or > minutes even though if you measure over a day it is accurate to less > than 1e-13. This is one problem with interpreting the "OSC" value. > > A second problem is that a TBolt can't really know its own accuracy; > that requires an external frequency reference, so take the PPS and OSC > with a grain of salt. However, in general, the TBolt will steer its 10 > MHz oscillator so as to minimize the PPS and OSC values. As such they > can be used as a rough idea of how well the unit is performing. > > /tvb > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
BC
Bob Camp
Sun, Sep 15, 2013 1:12 PM

Hi

The OCXO in the TBolt is quite low phase noise as a stand alone oscillator, It's not all that spur free in the TBolt. There's not a lot you can do about that with any settings. The OCXO in the NTBW50AA (and similar) is not as lot phase noise as the TBolt. It's also got issues "as installed". The whole settings thing does not apply to the NTBW50's since you can't change the loop settings on them with LH.

Bob

On Sep 15, 2013, at 7:14 AM, Jim Sanford wb4gcs@wb4gcs.org wrote:

So, what settings would you recommend for an application where frequency accuracy and low phase noise of the 10 MHz output are the objectives?

In particular, I'm not sure I understand how to establish the "extended TC method."

Thanks!
Jim
wb4gcs@amsat.org

On 9/14/2013 6:21 PM, WarrenS wrote:

OT for the current heading, so I renamed it "Oscillator stability"

as Tom says, "it's a complicated subject".
And to complicate things even further, here are a few of advanced subtleties that I've observed from the TBolt when using LadyHeather.

  1. The TBolt uses the received GPS signal as the reference for the calculated OSC freq offset and the Phase offset, over a measurement time of 1 sec.
    Unfortunately, from second to second the GPS signal is very noisy.
    Fortunately, LadyHeather can display plots of the data with several helpful user options, such as gain, offset, and most important length of time averaged.

  2. The Osc freq display plot of LH is so noisy that I find it mostly useless,
    unless I turn on LH's display filter to average the data over more than 1 second.
    LH allows you to manually turn-on and set the display filter to any time period desired,
    doing so this will greatly reduce the noise of the Osc plot.
    (10 samples is the display filter default, but I find 100sec to be a better place to start, for most things)
    A problem with the LH Osc freq plot even after filtering, is that the frequency difference data can not be counted on to be more accurate than a few parts in 1e-12 due to some offset rounding problems that often occurs.

  3. On the other hand, the filtered Phase plot has no known offset error.
    The Tbolt accurately shows time/phase different between the 1PPS and the received GPS signal,
    and when disciplined, it assumes if there is a difference, then the GPS is always right.
    This is why the antenna placement and setup is so important. Gunk in, Gunk out.

  4. The GPS signal, even on a 'perfect antenna', tends to wonders around ~10 ns PP independently of the time period averaged.
    So if LH is showing less than ~ 10 ns of GPS noise in the phase plot,
    it is because the control loop is set too fast and therefore forcing the Osc's freq to change a little so that it's phase will wonder around with the GPS's noise.
    Tbolt's max useable time constant is only 1000 sec, which is not nearly long enough to avoid this problem, when using a stable external osc like a good RB.
    To avoid tracking the noisy GPS data, the extended TC method must be used to set the Tbolt's osc to values >> 1000 seconds,
    and/or you can reduce the speed of  the phase tracking using the damping setting..

  5. How well and how fast the Tbolt minimizes the PPS phase and freq offset compared to the received GPS signal all depends on tuning.
    The Tbolt's TC determines what the Frequency  tracking time constant will be, and the TBolt's damping setting determines what the Phase tracking time multiplier is.
    If you set the damping factor to a large value like 100, then the Phase tracking will pretty much be turned off,
    making the disciplined loop more of a freq lock loop instead of a phase lock loop. This is done by lowering the gain of the loop's PID integrator.
    This is a way to set the phase's tracking time constant to be much slower than the freq TC setting, and if desired the phase tracking TC can be made several days long.
    Turning off the phase tracking has it's own set of pros and cons, but in most cases it is generally not desirerable in GPSDO.
    A damping setting of 0.7 to 1 will give the best overall compromise between the trade-off of not adding extra freq noise but still allowing good phase tracking.

  6. What some do not realize is to correct for any phase drift error, the Oscillator must be set off frequency.
    When the frequency is correct there is no further change in the present phase, whatever the present phase may be or wherever it may of come from.
    The faster you want to correct or change the present phase difference, no mater how it got there, the larger that the present frequency error must be made. (this causes freq noise)
    The trade off is, if you do not correct the present phase error then the past average frequency will be in error.

  7. So it is all a matter of what is more important to the application, present frequency error and noise or the average of all past frequency errors?
    The goal of most GPSDO is to keep the average past frequency error to zero. (a Phase Lock Loop)
    Where as for many transmitter things such as used by Hams,  it is the present errors and noise that is more important.
    So no need to cause a present freq error just to fix something that happened in the past.
    The past is gone and what happened before does not matter anymore. (a Freq Lock Loop)

ws


----- Original Message ----- From: "Tom Van Baak"
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Subject: Re: [time-nuts] New NTBW50AA

I get a spread of around 300ppt, that means I'm always within
300x10^-9 Hz of 10MHz or .0003Hz at 1GHz?

Note "300 ppt" is a "fractional frequency", unit-less value, so
at 10 MHz, 300e-12 * 1e7 Hz =  0.003 Hz
at 1 GHz, 300e-12 * 1e9 Hz = 0.3 Hz

I suppose the ppt spread is pretty much a function of how stable the osc is once
other factors like temp, antenna position, sat acquisition, etc, are optimized?

Yes and no; it's a complicated subject. For now, just two points:

A measure of frequency implies some measurement duration. A given TBolt may be off by more than 1e-11 in frequency over seconds or minutes even though if you measure over a day it is accurate to less than 1e-13. This is one problem with interpreting the "OSC" value.

A second problem is that a TBolt can't really know its own accuracy; that requires an external frequency reference, so take the PPS and OSC with a grain of salt. However, in general, the TBolt will steer its 10 MHz oscillator so as to minimize the PPS and OSC values. As such they can be used as a rough idea of how well the unit is performing.

/tvb


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi The OCXO in the TBolt is quite low phase noise as a stand alone oscillator, It's not all that spur free in the TBolt. There's not a lot you can do about that with any settings. The OCXO in the NTBW50AA (and similar) is not as lot phase noise as the TBolt. It's also got issues "as installed". The whole settings thing does not apply to the NTBW50's since you can't change the loop settings on them with LH. Bob On Sep 15, 2013, at 7:14 AM, Jim Sanford <wb4gcs@wb4gcs.org> wrote: > So, what settings would you recommend for an application where frequency accuracy and low phase noise of the 10 MHz output are the objectives? > > In particular, I'm not sure I understand how to establish the "extended TC method." > > Thanks! > Jim > wb4gcs@amsat.org > > On 9/14/2013 6:21 PM, WarrenS wrote: >> OT for the current heading, so I renamed it "Oscillator stability" >> >> as Tom says, "it's a complicated subject". >> And to complicate things even further, here are a few of advanced subtleties that I've observed from the TBolt when using LadyHeather. >> >> >> 1) The TBolt uses the received GPS signal as the reference for the calculated OSC freq offset and the Phase offset, over a measurement time of 1 sec. >> Unfortunately, from second to second the GPS signal is very noisy. >> Fortunately, LadyHeather can display plots of the data with several helpful user options, such as gain, offset, and most important length of time averaged. >> >> 2) The Osc freq display plot of LH is so noisy that I find it mostly useless, >> unless I turn on LH's display filter to average the data over more than 1 second. >> LH allows you to manually turn-on and set the display filter to any time period desired, >> doing so this will greatly reduce the noise of the Osc plot. >> (10 samples is the display filter default, but I find 100sec to be a better place to start, for most things) >> A problem with the LH Osc freq plot even after filtering, is that the frequency difference data can not be counted on to be more accurate than a few parts in 1e-12 due to some offset rounding problems that often occurs. >> >> >> 3) On the other hand, the filtered Phase plot has no known offset error. >> The Tbolt accurately shows time/phase different between the 1PPS and the received GPS signal, >> and when disciplined, it assumes if there is a difference, then the GPS is always right. >> This is why the antenna placement and setup is so important. Gunk in, Gunk out. >> >> >> 4) The GPS signal, even on a 'perfect antenna', tends to wonders around ~10 ns PP independently of the time period averaged. >> So if LH is showing less than ~ 10 ns of GPS noise in the phase plot, >> it is because the control loop is set too fast and therefore forcing the Osc's freq to change a little so that it's phase will wonder around with the GPS's noise. >> Tbolt's max useable time constant is only 1000 sec, which is not nearly long enough to avoid this problem, when using a stable external osc like a good RB. >> To avoid tracking the noisy GPS data, the extended TC method must be used to set the Tbolt's osc to values >> 1000 seconds, >> and/or you can reduce the speed of the phase tracking using the damping setting.. >> >> 4) How well and how fast the Tbolt minimizes the PPS phase and freq offset compared to the received GPS signal all depends on tuning. >> The Tbolt's TC determines what the Frequency tracking time constant will be, and the TBolt's damping setting determines what the Phase tracking time multiplier is. >> If you set the damping factor to a large value like 100, then the Phase tracking will pretty much be turned off, >> making the disciplined loop more of a freq lock loop instead of a phase lock loop. This is done by lowering the gain of the loop's PID integrator. >> This is a way to set the phase's tracking time constant to be much slower than the freq TC setting, and if desired the phase tracking TC can be made several days long. >> Turning off the phase tracking has it's own set of pros and cons, but in most cases it is generally not desirerable in GPSDO. >> A damping setting of 0.7 to 1 will give the best overall compromise between the trade-off of not adding extra freq noise but still allowing good phase tracking. >> >> >> 5) What some do not realize is to correct for any phase drift error, the Oscillator must be set off frequency. >> When the frequency is correct there is no further change in the present phase, whatever the present phase may be or wherever it may of come from. >> The faster you want to correct or change the present phase difference, no mater how it got there, the larger that the present frequency error must be made. (this causes freq noise) >> The trade off is, if you do not correct the present phase error then the past average frequency will be in error. >> >> 6) So it is all a matter of what is more important to the application, present frequency error and noise or the average of all past frequency errors? >> The goal of most GPSDO is to keep the average past frequency error to zero. (a Phase Lock Loop) >> Where as for many transmitter things such as used by Hams, it is the present errors and noise that is more important. >> So no need to cause a present freq error just to fix something that happened in the past. >> The past is gone and what happened before does not matter anymore. (a Freq Lock Loop) >> >> ws >> >> *********************** >> >> ----- Original Message ----- From: "Tom Van Baak" >> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com> >> Subject: Re: [time-nuts] New NTBW50AA >> >> >>> I get a spread of around 300ppt, that means I'm always within >>> 300x10^-9 Hz of 10MHz or .0003Hz at 1GHz? >> >> Note "300 ppt" is a "fractional frequency", unit-less value, so >> at 10 MHz, 300e-12 * 1e7 Hz = 0.003 Hz >> at 1 GHz, 300e-12 * 1e9 Hz = 0.3 Hz >> >>> I suppose the ppt spread is pretty much a function of how stable the osc is once >>> other factors like temp, antenna position, sat acquisition, etc, are optimized? >> >> Yes and no; it's a complicated subject. For now, just two points: >> >> A measure of frequency implies some measurement duration. A given TBolt may be off by more than 1e-11 in frequency over seconds or minutes even though if you measure over a day it is accurate to less than 1e-13. This is one problem with interpreting the "OSC" value. >> >> A second problem is that a TBolt can't really know its own accuracy; that requires an external frequency reference, so take the PPS and OSC with a grain of salt. However, in general, the TBolt will steer its 10 MHz oscillator so as to minimize the PPS and OSC values. As such they can be used as a rough idea of how well the unit is performing. >> >> /tvb >> >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BC
Bob Camp
Sun, Sep 15, 2013 1:52 PM

Hi

Should drink coffee first and then type messages..

On Sep 15, 2013, at 9:12 AM, Bob Camp lists@rtty.us wrote:

Hi

The OCXO in the TBolt is quite low phase noise as a stand alone oscillator, It's not all that spur free in the TBolt. There's not a lot you can do about that with any settings. The OCXO in the NTBW50AA (and similar) is not as lot

how about LOW phase noise

phase noise as the TBolt. It's also got issues "as installed". The whole settings thing does not apply to the NTBW50's since you can't change the loop settings on them with LH.

Bob

Your best bet for low phase noise at microwaves is still to lock up a 100 MHz VCXO and don't sweat the noise & spurs at lower frequencies.

Bob

On Sep 15, 2013, at 7:14 AM, Jim Sanford wb4gcs@wb4gcs.org wrote:

So, what settings would you recommend for an application where frequency accuracy and low phase noise of the 10 MHz output are the objectives?

In particular, I'm not sure I understand how to establish the "extended TC method."

Thanks!
Jim
wb4gcs@amsat.org

On 9/14/2013 6:21 PM, WarrenS wrote:

OT for the current heading, so I renamed it "Oscillator stability"

as Tom says, "it's a complicated subject".
And to complicate things even further, here are a few of advanced subtleties that I've observed from the TBolt when using LadyHeather.

  1. The TBolt uses the received GPS signal as the reference for the calculated OSC freq offset and the Phase offset, over a measurement time of 1 sec.
    Unfortunately, from second to second the GPS signal is very noisy.
    Fortunately, LadyHeather can display plots of the data with several helpful user options, such as gain, offset, and most important length of time averaged.

  2. The Osc freq display plot of LH is so noisy that I find it mostly useless,
    unless I turn on LH's display filter to average the data over more than 1 second.
    LH allows you to manually turn-on and set the display filter to any time period desired,
    doing so this will greatly reduce the noise of the Osc plot.
    (10 samples is the display filter default, but I find 100sec to be a better place to start, for most things)
    A problem with the LH Osc freq plot even after filtering, is that the frequency difference data can not be counted on to be more accurate than a few parts in 1e-12 due to some offset rounding problems that often occurs.

  3. On the other hand, the filtered Phase plot has no known offset error.
    The Tbolt accurately shows time/phase different between the 1PPS and the received GPS signal,
    and when disciplined, it assumes if there is a difference, then the GPS is always right.
    This is why the antenna placement and setup is so important. Gunk in, Gunk out.

  4. The GPS signal, even on a 'perfect antenna', tends to wonders around ~10 ns PP independently of the time period averaged.
    So if LH is showing less than ~ 10 ns of GPS noise in the phase plot,
    it is because the control loop is set too fast and therefore forcing the Osc's freq to change a little so that it's phase will wonder around with the GPS's noise.
    Tbolt's max useable time constant is only 1000 sec, which is not nearly long enough to avoid this problem, when using a stable external osc like a good RB.
    To avoid tracking the noisy GPS data, the extended TC method must be used to set the Tbolt's osc to values >> 1000 seconds,
    and/or you can reduce the speed of  the phase tracking using the damping setting..

  5. How well and how fast the Tbolt minimizes the PPS phase and freq offset compared to the received GPS signal all depends on tuning.
    The Tbolt's TC determines what the Frequency  tracking time constant will be, and the TBolt's damping setting determines what the Phase tracking time multiplier is.
    If you set the damping factor to a large value like 100, then the Phase tracking will pretty much be turned off,
    making the disciplined loop more of a freq lock loop instead of a phase lock loop. This is done by lowering the gain of the loop's PID integrator.
    This is a way to set the phase's tracking time constant to be much slower than the freq TC setting, and if desired the phase tracking TC can be made several days long.
    Turning off the phase tracking has it's own set of pros and cons, but in most cases it is generally not desirerable in GPSDO.
    A damping setting of 0.7 to 1 will give the best overall compromise between the trade-off of not adding extra freq noise but still allowing good phase tracking.

  6. What some do not realize is to correct for any phase drift error, the Oscillator must be set off frequency.
    When the frequency is correct there is no further change in the present phase, whatever the present phase may be or wherever it may of come from.
    The faster you want to correct or change the present phase difference, no mater how it got there, the larger that the present frequency error must be made. (this causes freq noise)
    The trade off is, if you do not correct the present phase error then the past average frequency will be in error.

  7. So it is all a matter of what is more important to the application, present frequency error and noise or the average of all past frequency errors?
    The goal of most GPSDO is to keep the average past frequency error to zero. (a Phase Lock Loop)
    Where as for many transmitter things such as used by Hams,  it is the present errors and noise that is more important.
    So no need to cause a present freq error just to fix something that happened in the past.
    The past is gone and what happened before does not matter anymore. (a Freq Lock Loop)

ws


----- Original Message ----- From: "Tom Van Baak"
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Subject: Re: [time-nuts] New NTBW50AA

I get a spread of around 300ppt, that means I'm always within
300x10^-9 Hz of 10MHz or .0003Hz at 1GHz?

Note "300 ppt" is a "fractional frequency", unit-less value, so
at 10 MHz, 300e-12 * 1e7 Hz =  0.003 Hz
at 1 GHz, 300e-12 * 1e9 Hz = 0.3 Hz

I suppose the ppt spread is pretty much a function of how stable the osc is once
other factors like temp, antenna position, sat acquisition, etc, are optimized?

Yes and no; it's a complicated subject. For now, just two points:

A measure of frequency implies some measurement duration. A given TBolt may be off by more than 1e-11 in frequency over seconds or minutes even though if you measure over a day it is accurate to less than 1e-13. This is one problem with interpreting the "OSC" value.

A second problem is that a TBolt can't really know its own accuracy; that requires an external frequency reference, so take the PPS and OSC with a grain of salt. However, in general, the TBolt will steer its 10 MHz oscillator so as to minimize the PPS and OSC values. As such they can be used as a rough idea of how well the unit is performing.

/tvb


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Should drink coffee first and *then* type messages.. On Sep 15, 2013, at 9:12 AM, Bob Camp <lists@rtty.us> wrote: > Hi > > The OCXO in the TBolt is quite low phase noise as a stand alone oscillator, It's not all that spur free in the TBolt. There's not a lot you can do about that with any settings. The OCXO in the NTBW50AA (and similar) is not as lot how about LOW phase noise > phase noise as the TBolt. It's also got issues "as installed". The whole settings thing does not apply to the NTBW50's since you can't change the loop settings on them with LH. > > Bob Your best bet for low phase noise at microwaves is still to lock up a 100 MHz VCXO and don't sweat the noise & spurs at lower frequencies. Bob > > On Sep 15, 2013, at 7:14 AM, Jim Sanford <wb4gcs@wb4gcs.org> wrote: > >> So, what settings would you recommend for an application where frequency accuracy and low phase noise of the 10 MHz output are the objectives? >> >> In particular, I'm not sure I understand how to establish the "extended TC method." >> >> Thanks! >> Jim >> wb4gcs@amsat.org >> >> On 9/14/2013 6:21 PM, WarrenS wrote: >>> OT for the current heading, so I renamed it "Oscillator stability" >>> >>> as Tom says, "it's a complicated subject". >>> And to complicate things even further, here are a few of advanced subtleties that I've observed from the TBolt when using LadyHeather. >>> >>> >>> 1) The TBolt uses the received GPS signal as the reference for the calculated OSC freq offset and the Phase offset, over a measurement time of 1 sec. >>> Unfortunately, from second to second the GPS signal is very noisy. >>> Fortunately, LadyHeather can display plots of the data with several helpful user options, such as gain, offset, and most important length of time averaged. >>> >>> 2) The Osc freq display plot of LH is so noisy that I find it mostly useless, >>> unless I turn on LH's display filter to average the data over more than 1 second. >>> LH allows you to manually turn-on and set the display filter to any time period desired, >>> doing so this will greatly reduce the noise of the Osc plot. >>> (10 samples is the display filter default, but I find 100sec to be a better place to start, for most things) >>> A problem with the LH Osc freq plot even after filtering, is that the frequency difference data can not be counted on to be more accurate than a few parts in 1e-12 due to some offset rounding problems that often occurs. >>> >>> >>> 3) On the other hand, the filtered Phase plot has no known offset error. >>> The Tbolt accurately shows time/phase different between the 1PPS and the received GPS signal, >>> and when disciplined, it assumes if there is a difference, then the GPS is always right. >>> This is why the antenna placement and setup is so important. Gunk in, Gunk out. >>> >>> >>> 4) The GPS signal, even on a 'perfect antenna', tends to wonders around ~10 ns PP independently of the time period averaged. >>> So if LH is showing less than ~ 10 ns of GPS noise in the phase plot, >>> it is because the control loop is set too fast and therefore forcing the Osc's freq to change a little so that it's phase will wonder around with the GPS's noise. >>> Tbolt's max useable time constant is only 1000 sec, which is not nearly long enough to avoid this problem, when using a stable external osc like a good RB. >>> To avoid tracking the noisy GPS data, the extended TC method must be used to set the Tbolt's osc to values >> 1000 seconds, >>> and/or you can reduce the speed of the phase tracking using the damping setting.. >>> >>> 4) How well and how fast the Tbolt minimizes the PPS phase and freq offset compared to the received GPS signal all depends on tuning. >>> The Tbolt's TC determines what the Frequency tracking time constant will be, and the TBolt's damping setting determines what the Phase tracking time multiplier is. >>> If you set the damping factor to a large value like 100, then the Phase tracking will pretty much be turned off, >>> making the disciplined loop more of a freq lock loop instead of a phase lock loop. This is done by lowering the gain of the loop's PID integrator. >>> This is a way to set the phase's tracking time constant to be much slower than the freq TC setting, and if desired the phase tracking TC can be made several days long. >>> Turning off the phase tracking has it's own set of pros and cons, but in most cases it is generally not desirerable in GPSDO. >>> A damping setting of 0.7 to 1 will give the best overall compromise between the trade-off of not adding extra freq noise but still allowing good phase tracking. >>> >>> >>> 5) What some do not realize is to correct for any phase drift error, the Oscillator must be set off frequency. >>> When the frequency is correct there is no further change in the present phase, whatever the present phase may be or wherever it may of come from. >>> The faster you want to correct or change the present phase difference, no mater how it got there, the larger that the present frequency error must be made. (this causes freq noise) >>> The trade off is, if you do not correct the present phase error then the past average frequency will be in error. >>> >>> 6) So it is all a matter of what is more important to the application, present frequency error and noise or the average of all past frequency errors? >>> The goal of most GPSDO is to keep the average past frequency error to zero. (a Phase Lock Loop) >>> Where as for many transmitter things such as used by Hams, it is the present errors and noise that is more important. >>> So no need to cause a present freq error just to fix something that happened in the past. >>> The past is gone and what happened before does not matter anymore. (a Freq Lock Loop) >>> >>> ws >>> >>> *********************** >>> >>> ----- Original Message ----- From: "Tom Van Baak" >>> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com> >>> Subject: Re: [time-nuts] New NTBW50AA >>> >>> >>>> I get a spread of around 300ppt, that means I'm always within >>>> 300x10^-9 Hz of 10MHz or .0003Hz at 1GHz? >>> >>> Note "300 ppt" is a "fractional frequency", unit-less value, so >>> at 10 MHz, 300e-12 * 1e7 Hz = 0.003 Hz >>> at 1 GHz, 300e-12 * 1e9 Hz = 0.3 Hz >>> >>>> I suppose the ppt spread is pretty much a function of how stable the osc is once >>>> other factors like temp, antenna position, sat acquisition, etc, are optimized? >>> >>> Yes and no; it's a complicated subject. For now, just two points: >>> >>> A measure of frequency implies some measurement duration. A given TBolt may be off by more than 1e-11 in frequency over seconds or minutes even though if you measure over a day it is accurate to less than 1e-13. This is one problem with interpreting the "OSC" value. >>> >>> A second problem is that a TBolt can't really know its own accuracy; that requires an external frequency reference, so take the PPS and OSC with a grain of salt. However, in general, the TBolt will steer its 10 MHz oscillator so as to minimize the PPS and OSC values. As such they can be used as a rough idea of how well the unit is performing. >>> >>> /tvb >>> >>> >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
Q
quartz55
Sun, Sep 15, 2013 2:32 PM

The only other place I can put it would be about 100' to the south on the fence line, and unfortunately it's so steep, I've lost about 30' altitude there already.  So the farther away I get from the house to a clearer location, the lower it gets; in about 200 yards it drops 200' to Buck Run. And I think that's a loosing proposition.

I'll see if I can't get someone to help and install it on the rotating mast in line with the pipe and install the lower outrigger which will minimize the leveraged wind displacement.  Anyone nearby?  I can trade some UHF/VHF Spectras for help with the work.  I've got a few other things like mobile antennas that I can loose.  I may call up the local ARC.  I'm just real fearful of standing on that chimney any more, I don't mind walking around on the 4/12 part of the roof.

Almost 1/2 way through the SP command.

Dave
----- Original Message -----
From: Bob Camp
To: Discussion of precise time and frequency measurement
Sent: Saturday, September 14, 2013 1:40 PM
Subject: Re: [time-nuts] New NTBW50AA

Hi

As long as the wind does not blow the tower around, that's probably your best location.

Bob

The only other place I can put it would be about 100' to the south on the fence line, and unfortunately it's so steep, I've lost about 30' altitude there already. So the farther away I get from the house to a clearer location, the lower it gets; in about 200 yards it drops 200' to Buck Run. And I think that's a loosing proposition. I'll see if I can't get someone to help and install it on the rotating mast in line with the pipe and install the lower outrigger which will minimize the leveraged wind displacement. Anyone nearby? I can trade some UHF/VHF Spectras for help with the work. I've got a few other things like mobile antennas that I can loose. I may call up the local ARC. I'm just real fearful of standing on that chimney any more, I don't mind walking around on the 4/12 part of the roof. Almost 1/2 way through the SP command. Dave ----- Original Message ----- From: Bob Camp To: Discussion of precise time and frequency measurement Sent: Saturday, September 14, 2013 1:40 PM Subject: Re: [time-nuts] New NTBW50AA Hi As long as the wind does not blow the tower around, that's probably your best location. Bob
BS
Bob Stewart
Sun, Sep 15, 2013 2:40 PM

It seems to me that you're fixing a problem that doesn't exist.  Have you tried just putting it somewhere convenient and leaving it alone to see what happens?

Bob - AE6RV


From: quartz55 quartz55@hughes.net
To: Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Sunday, September 15, 2013 9:32 AM
Subject: Re: [time-nuts] New NTBW50AA

The only other place I can put it would be about 100' to the south on the fence line, and unfortunately it's so steep, I've lost about 30' altitude there already.  So the farther away I get from the house to a clearer location, the lower it gets; in about 200 yards it drops 200' to Buck Run. And I think that's a loosing proposition.

I'll see if I can't get someone to help and install it on the rotating mast in line with the pipe and install the lower outrigger which will minimize the leveraged wind displacement.  Anyone nearby?  I can trade some UHF/VHF Spectras for help with the work.  I've got a few other things like mobile antennas that I can loose.  I may call up the local ARC.  I'm just real fearful of standing on that chimney any more, I don't mind walking around on the 4/12 part of the roof.

Almost 1/2 way through the SP command.

Dave
  ----- Original Message -----
  From: Bob Camp
  To: Discussion of precise time and frequency measurement
  Sent: Saturday, September 14, 2013 1:40 PM
  Subject: Re: [time-nuts] New NTBW50AA

  Hi

  As long as the wind does not blow the tower around, that's probably your best location.

  Bob


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

It seems to me that you're fixing a problem that doesn't exist.  Have you tried just putting it somewhere convenient and leaving it alone to see what happens? Bob - AE6RV >________________________________ > From: quartz55 <quartz55@hughes.net> >To: Discussion of precise time and frequency measurement <time-nuts@febo.com> >Sent: Sunday, September 15, 2013 9:32 AM >Subject: Re: [time-nuts] New NTBW50AA > > >The only other place I can put it would be about 100' to the south on the fence line, and unfortunately it's so steep, I've lost about 30' altitude there already.  So the farther away I get from the house to a clearer location, the lower it gets; in about 200 yards it drops 200' to Buck Run. And I think that's a loosing proposition. > >I'll see if I can't get someone to help and install it on the rotating mast in line with the pipe and install the lower outrigger which will minimize the leveraged wind displacement.  Anyone nearby?  I can trade some UHF/VHF Spectras for help with the work.  I've got a few other things like mobile antennas that I can loose.  I may call up the local ARC.  I'm just real fearful of standing on that chimney any more, I don't mind walking around on the 4/12 part of the roof. > >Almost 1/2 way through the SP command. > >Dave >  ----- Original Message ----- >  From: Bob Camp >  To: Discussion of precise time and frequency measurement >  Sent: Saturday, September 14, 2013 1:40 PM >  Subject: Re: [time-nuts] New NTBW50AA > > >  Hi > >  As long as the wind does not blow the tower around, that's probably your best location. > >  Bob >_______________________________________________ >time-nuts mailing list -- time-nuts@febo.com >To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >and follow the instructions there. > > >
CS
Charles Steinmetz
Sun, Sep 15, 2013 4:53 PM

Dave wrote:

The only other place I can put it would be about 100' to the south
on the fence line, and unfortunately it's so steep, I've lost about
30' altitude there already.  So the farther away I get from the
house to a clearer location, the lower it gets; in about 200 yards
it drops 200'

The height of the antenna above average terrain doesn't mean a lot
with satellite signals.  What you need is a clear view of the sky
above 10 or 15 degrees elevation for as much of the azimuth as you
can arrange (and particularly the southern hemisphere of sky).  So,
for example, if you have a "high spot" on the property that is
covered with trees that you can't get above with your GPS antenna,
and a lower spot with a clear sky view, the lower spot is better for
a GPS antenna.  A hillside that slopes off to the south can be an
excellent location.  It sounds like somewhere to the south toward or
at the fence line may be your best location.  Use direct-burial coax,
rent a ditch witch for the day, and you're set.  Remember -- it is
clear sky view you are after, NOT height.

A bit of height above the nearby ground can help with
multipath.  Survey-grade choke-ring antennas are better about this by
design.  Some folks use a metal "ground plane" under the GPS antenna
(it is not really operating as a ground plane, rather, it is simply
shielding the antenna from low-incidence signals).  If the antenna is
at least 10 feet above the nearby ground and your elevation mask is
set to 15 degrees or more, you shouldn't have multipath problems
serious enough to degrade GPS timing (i.e., other errors will
dominate).  Note that the elevation mask can't reject the multipath
itself -- it doesn't steer the antenna sensitivity -- but low
satellites are the ones that suffer the most multipath unless you are
in an "urban canyon" or have other strong L-band reflectors nearby,
so by excluding low satellites you also exclude the multipath.  Since
low satellites also suffer the worst atmospheric variability,
excluding them generally tightens up the timing solution also.

Best regards,

Charles

Dave wrote: >The only other place I can put it would be about 100' to the south >on the fence line, and unfortunately it's so steep, I've lost about >30' altitude there already. So the farther away I get from the >house to a clearer location, the lower it gets; in about 200 yards >it drops 200' The height of the antenna above average terrain doesn't mean a lot with satellite signals. What you need is a clear view of the sky above 10 or 15 degrees elevation for as much of the azimuth as you can arrange (and particularly the southern hemisphere of sky). So, for example, if you have a "high spot" on the property that is covered with trees that you can't get above with your GPS antenna, and a lower spot with a clear sky view, the lower spot is better for a GPS antenna. A hillside that slopes off to the south can be an excellent location. It sounds like somewhere to the south toward or at the fence line may be your best location. Use direct-burial coax, rent a ditch witch for the day, and you're set. Remember -- it is clear sky view you are after, NOT height. A bit of height above the nearby ground can help with multipath. Survey-grade choke-ring antennas are better about this by design. Some folks use a metal "ground plane" under the GPS antenna (it is not really operating as a ground plane, rather, it is simply shielding the antenna from low-incidence signals). If the antenna is at least 10 feet above the nearby ground and your elevation mask is set to 15 degrees or more, you shouldn't have multipath problems serious enough to degrade GPS timing (i.e., other errors will dominate). Note that the elevation mask can't reject the multipath itself -- it doesn't steer the antenna sensitivity -- but low satellites are the ones that suffer the most multipath unless you are in an "urban canyon" or have other strong L-band reflectors nearby, so by excluding low satellites you also exclude the multipath. Since low satellites also suffer the worst atmospheric variability, excluding them generally tightens up the timing solution also. Best regards, Charles
Q
quartz55
Sun, Sep 15, 2013 6:13 PM

Well, I could try it down on the fence, but like I say when I get down there, I'm not sure the visibility above the trees is any better to the east and west because it's so much lower and the trees are all 80' high or more.  Add another 30-40' for the drop and you can see what I mean.  Like I say, it's a 200' drop from the house to the barn which is only 120 yards away.  I'm not thrilled having it above the beams, it's the highest metal thing in the area and most likely more prone to lightning than anything else excpet the trees and it will blow around a bit up there as well as rotating once in a while.  I did have a strike here once, but it was down by the barn and was looking for the underground telephone cables, it blew apart a post and knocked off the boards, it was maybe 100 yards from the shack.  It wouldn't be much of a problem to dig a tiny trench to the fence and I may try that after this precise survey is done, then do another and see if anything changes.  I could hide it in a bird box at the fence too.  After all at this point I'm just playing.  I'll eventually make up my mind.  Also, I don't have an attic, so that's not an option.

Dave
N3DT

Well, I could try it down on the fence, but like I say when I get down there, I'm not sure the visibility above the trees is any better to the east and west because it's so much lower and the trees are all 80' high or more. Add another 30-40' for the drop and you can see what I mean. Like I say, it's a 200' drop from the house to the barn which is only 120 yards away. I'm not thrilled having it above the beams, it's the highest metal thing in the area and most likely more prone to lightning than anything else excpet the trees and it will blow around a bit up there as well as rotating once in a while. I did have a strike here once, but it was down by the barn and was looking for the underground telephone cables, it blew apart a post and knocked off the boards, it was maybe 100 yards from the shack. It wouldn't be much of a problem to dig a tiny trench to the fence and I may try that after this precise survey is done, then do another and see if anything changes. I could hide it in a bird box at the fence too. After all at this point I'm just playing. I'll eventually make up my mind. Also, I don't have an attic, so that's not an option. Dave N3DT
CS
Charles Steinmetz
Sun, Sep 15, 2013 7:04 PM

Dave wrote:

Well, I could try it down on the fence, but like I say when I get
down there, I'm not sure the visibility above the trees is any
better to the east and west because it's so much lower and the trees
are all 80' high or more.

It's not just the height of the trees, it's the horizontal distance
from the antenna location to the trees, too -- the elevation angle
above the antenna location (elevation = arctan (tree height/distance
to trees) (in this case, "tree height" means height of the trees
above the horizontal plane of the antenna, so it includes any
difference in the ground height at the tree location, as well).

At the end of the day, you want the most open sky you can get,
particularly the southern hemisphere of sky, but the lowest
elevations (say, under 20 degrees elevation) are not nearly as
important as 20 through 90 degrees.

Best regards,

Charles

Dave wrote: >Well, I could try it down on the fence, but like I say when I get >down there, I'm not sure the visibility above the trees is any >better to the east and west because it's so much lower and the trees >are all 80' high or more. It's not just the height of the trees, it's the horizontal distance from the antenna location to the trees, too -- the elevation angle above the antenna location (elevation = arctan (tree height/distance to trees) (in this case, "tree height" means height of the trees above the horizontal plane of the antenna, so it includes any difference in the ground height at the tree location, as well). At the end of the day, you want the most open sky you can get, particularly the southern hemisphere of sky, but the lowest elevations (say, under 20 degrees elevation) are not nearly as important as 20 through 90 degrees. Best regards, Charles
Q
quartz55
Sun, Sep 15, 2013 8:11 PM

Yes, I know it's not just the height of the trees, but the distance away and I'm really only getting 50-60' away at most and the hill is steep, it's about 20° in the worst spots.  This is not your flat suburban lot, it's in the foothills of the Blue Ridge.  The only flat spots are where we've made them.  I'll have to go out and do a visual survey to see which one actually gives more free sky from say 20° up.  Right now I'm tending towards the one towards the fence because it's not on the beam mast.  But I'll let this survey finish.  The WX is going to be good enough the rest of the week it'll be good outside working WX.

It looks like the fence is really only 15' lower, if I put it at the arrow between those 2 gardens, I can keep the coax run at 100' and still have it up maybe 15'. http://i251.photobucket.com/albums/gg287/DogTi/time/land_zps9ee07b7a.jpg

Dave
It's not just the height of the trees, it's the horizontal distance
from the antenna location to the trees, too -- the elevation angle
above the antenna location (elevation = arctan (tree height/distance
to trees) (in this case, "tree height" means height of the trees
above the horizontal plane of the antenna, so it includes any
difference in the ground height at the tree location, as well).

At the end of the day, you want the most open sky you can get,
particularly the southern hemisphere of sky, but the lowest
elevations (say, under 20 degrees elevation) are not nearly as
important as 20 through 90 degrees.

Best regards,

Charles


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1432 / Virus Database: 3222/6167 - Release Date: 09/15/13

Yes, I know it's not just the height of the trees, but the distance away and I'm really only getting 50-60' away at most and the hill is steep, it's about 20° in the worst spots. This is not your flat suburban lot, it's in the foothills of the Blue Ridge. The only flat spots are where we've made them. I'll have to go out and do a visual survey to see which one actually gives more free sky from say 20° up. Right now I'm tending towards the one towards the fence because it's not on the beam mast. But I'll let this survey finish. The WX is going to be good enough the rest of the week it'll be good outside working WX. It looks like the fence is really only 15' lower, if I put it at the arrow between those 2 gardens, I can keep the coax run at 100' and still have it up maybe 15'. http://i251.photobucket.com/albums/gg287/DogTi/time/land_zps9ee07b7a.jpg Dave It's not just the height of the trees, it's the horizontal distance from the antenna location to the trees, too -- the elevation angle above the antenna location (elevation = arctan (tree height/distance to trees) (in this case, "tree height" means height of the trees above the horizontal plane of the antenna, so it includes any difference in the ground height at the tree location, as well). At the end of the day, you want the most open sky you can get, particularly the southern hemisphere of sky, but the lowest elevations (say, under 20 degrees elevation) are not nearly as important as 20 through 90 degrees. Best regards, Charles _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ----- No virus found in this message. Checked by AVG - www.avg.com Version: 10.0.1432 / Virus Database: 3222/6167 - Release Date: 09/15/13
DL
Don Latham
Sun, Sep 15, 2013 9:54 PM

Lee, if I can make a couple of comments, first, you can have large
voltage differences among your 23 "ground" rods, depending on the ground
terminus of the lightning strokes. Second, antennae and feed lines that
do not attract lightning due to being "hidden" can still be subject to
large induced voltages.
I'd suggest all the ground rods and antenna bases be connected directly
with large cable so they are all truly at the same ground potential...
Don

Lee Mushel

Lightning is certainly one of Old Mother Nature's principal weapons as
well
as being extremely unpredictable.  My House/Shack is on the side of a
hill
with the Antenna Farm on top of the hill about 120 yds away with an
additional 30 ft. of elevation.  There is a metal "lightning rod"
extending
high enough to provide a "cone of protection" for the house and is
grounded
directly.  And there are five other metal antennas on that roof, two to
a
separate ground and three to still another which is close to the shack
"ground field" which is three additional  ground rods.  All ground rods
I
mention are the standard 8 ft. copper plated steel items.  Each of five
antennas in the antenna field also have individual ground rods and I
guess
if you follow the cable shields you could say that they are all
connected
together.    A couple of years ago I counted all ground rods and came up
with 23.  Before all of the metal was added and the "lightning rod" was
the
only metal extending about 20 ft. above the roof and we had an
electrical
storm my wife would complain that she could feel her hair trying to rise
in
response to the field coming from the cable which ran across the roof
and at
it's closest point was only 6-7 ft. from her head as the crow flies.
But
after all the additional metal was added in the form of antennas with
grounds  she no longer complained.

I do not claim to understand what goes on around here at an elevation of
about 102 ft. above the valley floor but I believe that we have been
"fortunate"  (or lucky if you prefer that word) and we have so much
metal in
the air that most "discharges" are not of the type that is accompanied
by
thunder!  Two years ago we had a tree die that was no more than 30 ft.
from
the house.  Since I have not always been able to accurately predict
where
trees will fall that I cut myself I hired a professional to take it down
and
he took one look at it, reached up and stripped off a piece of bark and
commented, "See those lines?  This tree was struck by lightning.
That's
what killed it!"  But no lightning related damage to structure or
equipment
here for nearly 15 years!

My only thought on the subject was that only a fool would fail to
disconnect
an antenna that rises above the house roof  during an electrical storm.
I
freely admit that I have several ham radio VHF antennas  (and a GPS
antenna
supplying the disciplined system for frequency reference) well below
roof
peak level  (clear view to south for GPS, of course) that I do not
disconnect from inexpensive radios maintained for the purpose.

This is risk I accept.  And I certainly have seen paths followed by
lightning that are truly amazing!  ....and horrendously damaging.  Never
forget that you only live once and you do have a perfect right to enjoy
your
hobby without fear.

Regards,

Lee A. Mushel  K9WRU

----- Original Message -----
From: "quartz55" quartz55@hughes.net
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Sent: Sunday, September 15, 2013 12:13 PM
Subject: Re: [time-nuts] New NTBW50AA

.  I'm not thrilled having it above the beams, it's the highest metal
thing
in the area and most likely more prone to lightning than anything else
excpet the trees and it will blow around a bit up there as well as
rotating
once in a while.  I did have a strike here once, but it was down by the
barn
and was looking for the underground telephone cables, it blew apart a
post
and knocked off the boards, it was maybe 100 yards from the shack.  It
wouldn't be much of a problem to dig a tiny trench to the fence and I
may
try that after this precise survey is done, then do another and see if
anything changes.  I could hide it i

n a bird box at the fence too.  After all at this point I'm just
playing.
I'll eventually make up my mind.  Also, I don't have an attic, so
that's
not an option.

Dave
N3DT


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
“The power of accurate observation is commonly called cynicism by those
who have not got it.”
-George Bernard Shaw

Dr. Don Latham AJ7LL
Six Mile Systems LLC
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
Skype: buffler2
www.lightningforensics.com
www.sixmilesystems.com

Lee, if I can make a couple of comments, first, you can have large voltage differences among your 23 "ground" rods, depending on the ground terminus of the lightning strokes. Second, antennae and feed lines that do not attract lightning due to being "hidden" can still be subject to large induced voltages. I'd suggest all the ground rods and antenna bases be connected directly with large cable so they are all truly at the same ground potential... Don Lee Mushel > Lightning is certainly one of Old Mother Nature's principal weapons as > well > as being extremely unpredictable. My House/Shack is on the side of a > hill > with the Antenna Farm on top of the hill about 120 yds away with an > additional 30 ft. of elevation. There is a metal "lightning rod" > extending > high enough to provide a "cone of protection" for the house and is > grounded > directly. And there are five other metal antennas on that roof, two to > a > separate ground and three to still another which is close to the shack > "ground field" which is three additional ground rods. All ground rods > I > mention are the standard 8 ft. copper plated steel items. Each of five > antennas in the antenna field also have individual ground rods and I > guess > if you follow the cable shields you could say that they are all > connected > together. A couple of years ago I counted all ground rods and came up > with 23. Before all of the metal was added and the "lightning rod" was > the > only metal extending about 20 ft. above the roof and we had an > electrical > storm my wife would complain that she could feel her hair trying to rise > in > response to the field coming from the cable which ran across the roof > and at > it's closest point was only 6-7 ft. from her head as the crow flies. > But > after all the additional metal was added in the form of antennas with > grounds she no longer complained. > > I do not claim to understand what goes on around here at an elevation of > about 102 ft. above the valley floor but I believe that we have been > "fortunate" (or lucky if you prefer that word) and we have so much > metal in > the air that most "discharges" are not of the type that is accompanied > by > thunder! Two years ago we had a tree die that was no more than 30 ft. > from > the house. Since I have not always been able to accurately predict > where > trees will fall that I cut myself I hired a professional to take it down > and > he took one look at it, reached up and stripped off a piece of bark and > commented, "See those lines? This tree was struck by lightning. > That's > what killed it!" But no lightning related damage to structure or > equipment > here for nearly 15 years! > > My only thought on the subject was that only a fool would fail to > disconnect > an antenna that rises above the house roof during an electrical storm. > I > freely admit that I have several ham radio VHF antennas (and a GPS > antenna > supplying the disciplined system for frequency reference) well below > roof > peak level (clear view to south for GPS, of course) that I do not > disconnect from inexpensive radios maintained for the purpose. > > This is risk I accept. And I certainly have seen paths followed by > lightning that are truly amazing! ....and horrendously damaging. Never > forget that you only live once and you do have a perfect right to enjoy > your > hobby without fear. > > Regards, > > Lee A. Mushel K9WRU > > > ----- Original Message ----- > From: "quartz55" <quartz55@hughes.net> > To: "Discussion of precise time and frequency measurement" > <time-nuts@febo.com> > Sent: Sunday, September 15, 2013 12:13 PM > Subject: Re: [time-nuts] New NTBW50AA > > > . I'm not thrilled having it above the beams, it's the highest metal > thing > in the area and most likely more prone to lightning than anything else > excpet the trees and it will blow around a bit up there as well as > rotating > once in a while. I did have a strike here once, but it was down by the > barn > and was looking for the underground telephone cables, it blew apart a > post > and knocked off the boards, it was maybe 100 yards from the shack. It > wouldn't be much of a problem to dig a tiny trench to the fence and I > may > try that after this precise survey is done, then do another and see if > anything changes. I could hide it i >> n a bird box at the fence too. After all at this point I'm just >> playing. >> I'll eventually make up my mind. Also, I don't have an attic, so >> that's >> not an option. >> >> Dave >> N3DT >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> >> > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > -- “The power of accurate observation is commonly called cynicism by those who have not got it.” -George Bernard Shaw Dr. Don Latham AJ7LL Six Mile Systems LLC 17850 Six Mile Road POB 134 Huson, MT, 59846 VOX 406-626-4304 Skype: buffler2 www.lightningforensics.com www.sixmilesystems.com
LM
Lee Mushel
Sun, Sep 15, 2013 10:16 PM

Lightning is certainly one of Old Mother Nature's principal weapons as well
as being extremely unpredictable.  My House/Shack is on the side of a hill
with the Antenna Farm on top of the hill about 120 yds away with an
additional 30 ft. of elevation.  There is a metal "lightning rod" extending
high enough to provide a "cone of protection" for the house and is grounded
directly.  And there are five other metal antennas on that roof, two to a
separate ground and three to still another which is close to the shack
"ground field" which is three additional  ground rods.  All ground rods I
mention are the standard 8 ft. copper plated steel items.  Each of five
antennas in the antenna field also have individual ground rods and I guess
if you follow the cable shields you could say that they are all connected
together.    A couple of years ago I counted all ground rods and came up
with 23.  Before all of the metal was added and the "lightning rod" was the
only metal extending about 20 ft. above the roof and we had an electrical
storm my wife would complain that she could feel her hair trying to rise in
response to the field coming from the cable which ran across the roof and at
it's closest point was only 6-7 ft. from her head as the crow flies.  But
after all the additional metal was added in the form of antennas with
grounds  she no longer complained.

I do not claim to understand what goes on around here at an elevation of
about 102 ft. above the valley floor but I believe that we have been
"fortunate"  (or lucky if you prefer that word) and we have so much metal in
the air that most "discharges" are not of the type that is accompanied by
thunder!  Two years ago we had a tree die that was no more than 30 ft. from
the house.  Since I have not always been able to accurately predict where
trees will fall that I cut myself I hired a professional to take it down and
he took one look at it, reached up and stripped off a piece of bark and
commented, "See those lines?  This tree was struck by lightning.  That's
what killed it!"  But no lightning related damage to structure or equipment
here for nearly 15 years!

My only thought on the subject was that only a fool would fail to disconnect
an antenna that rises above the house roof  during an electrical storm.  I
freely admit that I have several ham radio VHF antennas  (and a GPS antenna
supplying the disciplined system for frequency reference) well below roof
peak level  (clear view to south for GPS, of course) that I do not
disconnect from inexpensive radios maintained for the purpose.

This is risk I accept.  And I certainly have seen paths followed by
lightning that are truly amazing!  ....and horrendously damaging.  Never
forget that you only live once and you do have a perfect right to enjoy your
hobby without fear.

Regards,

Lee A. Mushel  K9WRU

----- Original Message -----
From: "quartz55" quartz55@hughes.net
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Sent: Sunday, September 15, 2013 12:13 PM
Subject: Re: [time-nuts] New NTBW50AA

.  I'm not thrilled having it above the beams, it's the highest metal thing
in the area and most likely more prone to lightning than anything else
excpet the trees and it will blow around a bit up there as well as rotating
once in a while.  I did have a strike here once, but it was down by the barn
and was looking for the underground telephone cables, it blew apart a post
and knocked off the boards, it was maybe 100 yards from the shack.  It
wouldn't be much of a problem to dig a tiny trench to the fence and I may
try that after this precise survey is done, then do another and see if
anything changes.  I could hide it i

n a bird box at the fence too.  After all at this point I'm just playing.
I'll eventually make up my mind.  Also, I don't have an attic, so that's
not an option.

Dave
N3DT


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Lightning is certainly one of Old Mother Nature's principal weapons as well as being extremely unpredictable. My House/Shack is on the side of a hill with the Antenna Farm on top of the hill about 120 yds away with an additional 30 ft. of elevation. There is a metal "lightning rod" extending high enough to provide a "cone of protection" for the house and is grounded directly. And there are five other metal antennas on that roof, two to a separate ground and three to still another which is close to the shack "ground field" which is three additional ground rods. All ground rods I mention are the standard 8 ft. copper plated steel items. Each of five antennas in the antenna field also have individual ground rods and I guess if you follow the cable shields you could say that they are all connected together. A couple of years ago I counted all ground rods and came up with 23. Before all of the metal was added and the "lightning rod" was the only metal extending about 20 ft. above the roof and we had an electrical storm my wife would complain that she could feel her hair trying to rise in response to the field coming from the cable which ran across the roof and at it's closest point was only 6-7 ft. from her head as the crow flies. But after all the additional metal was added in the form of antennas with grounds she no longer complained. I do not claim to understand what goes on around here at an elevation of about 102 ft. above the valley floor but I believe that we have been "fortunate" (or lucky if you prefer that word) and we have so much metal in the air that most "discharges" are not of the type that is accompanied by thunder! Two years ago we had a tree die that was no more than 30 ft. from the house. Since I have not always been able to accurately predict where trees will fall that I cut myself I hired a professional to take it down and he took one look at it, reached up and stripped off a piece of bark and commented, "See those lines? This tree was struck by lightning. That's what killed it!" But no lightning related damage to structure or equipment here for nearly 15 years! My only thought on the subject was that only a fool would fail to disconnect an antenna that rises above the house roof during an electrical storm. I freely admit that I have several ham radio VHF antennas (and a GPS antenna supplying the disciplined system for frequency reference) well below roof peak level (clear view to south for GPS, of course) that I do not disconnect from inexpensive radios maintained for the purpose. This is risk I accept. And I certainly have seen paths followed by lightning that are truly amazing! ....and horrendously damaging. Never forget that you only live once and you do have a perfect right to enjoy your hobby without fear. Regards, Lee A. Mushel K9WRU ----- Original Message ----- From: "quartz55" <quartz55@hughes.net> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com> Sent: Sunday, September 15, 2013 12:13 PM Subject: Re: [time-nuts] New NTBW50AA . I'm not thrilled having it above the beams, it's the highest metal thing in the area and most likely more prone to lightning than anything else excpet the trees and it will blow around a bit up there as well as rotating once in a while. I did have a strike here once, but it was down by the barn and was looking for the underground telephone cables, it blew apart a post and knocked off the boards, it was maybe 100 yards from the shack. It wouldn't be much of a problem to dig a tiny trench to the fence and I may try that after this precise survey is done, then do another and see if anything changes. I could hide it i > n a bird box at the fence too. After all at this point I'm just playing. > I'll eventually make up my mind. Also, I don't have an attic, so that's > not an option. > > Dave > N3DT > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > >
PB
Paul Berger
Mon, Sep 16, 2013 12:28 AM

All of this talk about antenna position has got me thinking about a
better place for my antenna.  I live in a rental townhouse and have no
access to the roof and even if I did bringing a cable in from outside
would be difficult, so  am pretty much limited to an indoor antenna.  My
current antenna location is in  a window on the second floor facing
north west at a bearing of about 300 degrees, about 15 ft above the
ground, as as many of you might guess it is not working that well.

Looking around one possible other location would be on the top floor
about 10 ft higher , with a straight out bearing if about 120 degrees on
an exterior wall. The building is wood frame with drywall on the inside,
fibreglass insulation and the exterior is covered with vinyl siding.
Ideally I would like to stand the antenna in the corner of the room, but
if it would make any difference there is a window in the room I could
place the antenna in front of.  This location will place it nearly as
high as the surrounding roof tops and beyond that there is nothing as I
an near the height of land in the area and a couple miles to the south
is the Atlantic Ocean and its all water after that.  The townhouse was
pre-wired with RG-6 coax and I could use that to bring the signal to
where my Nortel unit is.  From what I gather RG-6 cable will work fine
but I have a couple of concerns.

  1. To get to where the receiver is I will need to go through the cable
    drop in the room where I plan to put the antenna, to the basement where
    all the drops are terminated, and then back up on another drop, so I
    will end up with 4 segments of cable and three couplings, will these
    couplings be and issue?  The connectors involved would be F connectors.

  2. At the antenna end it would be more convenient to use a piece of
    RG-59 cable between the antenna and the wall since it already has a TNC
    connector on the end that I need for the antenna, since they are both 75
    ohm cables I don't think it would be an issue but I don't know for sure.

One final question when we set a filter for elevation is that the
elevation above ground or the elevation above the plane of the antenna?

Paul.

All of this talk about antenna position has got me thinking about a better place for my antenna. I live in a rental townhouse and have no access to the roof and even if I did bringing a cable in from outside would be difficult, so am pretty much limited to an indoor antenna. My current antenna location is in a window on the second floor facing north west at a bearing of about 300 degrees, about 15 ft above the ground, as as many of you might guess it is not working that well. Looking around one possible other location would be on the top floor about 10 ft higher , with a straight out bearing if about 120 degrees on an exterior wall. The building is wood frame with drywall on the inside, fibreglass insulation and the exterior is covered with vinyl siding. Ideally I would like to stand the antenna in the corner of the room, but if it would make any difference there is a window in the room I could place the antenna in front of. This location will place it nearly as high as the surrounding roof tops and beyond that there is nothing as I an near the height of land in the area and a couple miles to the south is the Atlantic Ocean and its all water after that. The townhouse was pre-wired with RG-6 coax and I could use that to bring the signal to where my Nortel unit is. From what I gather RG-6 cable will work fine but I have a couple of concerns. 1. To get to where the receiver is I will need to go through the cable drop in the room where I plan to put the antenna, to the basement where all the drops are terminated, and then back up on another drop, so I will end up with 4 segments of cable and three couplings, will these couplings be and issue? The connectors involved would be F connectors. 2. At the antenna end it would be more convenient to use a piece of RG-59 cable between the antenna and the wall since it already has a TNC connector on the end that I need for the antenna, since they are both 75 ohm cables I don't think it would be an issue but I don't know for sure. One final question when we set a filter for elevation is that the elevation above ground or the elevation above the plane of the antenna? Paul.
LM
Lee Mushel
Mon, Sep 16, 2013 2:25 AM

Don,

Thanks for the suggestion!  I have no doubt that you are correct and
especially so at the moment of any "strike" or discharge.  But these rods
cover an area approximately 200 by 500 ft. and I don't find heavy cable
"practical!"  Feedline varies from the RG-58 you would expect to 7/8 inch
Andrew (350').

Lee
----- Original Message -----
From: "Don Latham" djl@montana.com
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Sent: Sunday, September 15, 2013 3:54 PM
Subject: Re: [time-nuts] New NTBW50AA

Lee, if I can make a couple of comments, first, you can have large
voltage differences among your 23 "ground" rods, depending on the ground
terminus of the lightning strokes. Second, antennae and feed lines that
do not attract lightning due to being "hidden" can still be subject to
large induced voltages.
I'd suggest all the ground rods and antenna bases be connected directly
with large cable so they are all truly at the same ground potential...
Don

Lee Mushel

Lightning is certainly one of Old Mother Nature's principal weapons as
well
as being extremely unpredictable.  My House/Shack is on the side of a
hill
with the Antenna Farm on top of the hill about 120 yds away with an
additional 30 ft. of elevation.  There is a metal "lightning rod"
extending
high enough to provide a "cone of protection" for the house and is
grounded
directly.  And there are five other metal antennas on that roof, two to
a
separate ground and three to still another which is close to the shack
"ground field" which is three additional  ground rods.  All ground rods
I
mention are the standard 8 ft. copper plated steel items.  Each of five
antennas in the antenna field also have individual ground rods and I
guess
if you follow the cable shields you could say that they are all
connected
together.    A couple of years ago I counted all ground rods and came up
with 23.  Before all of the metal was added and the "lightning rod" was
the
only metal extending about 20 ft. above the roof and we had an
electrical
storm my wife would complain that she could feel her hair trying to rise
in
response to the field coming from the cable which ran across the roof
and at
it's closest point was only 6-7 ft. from her head as the crow flies.
But
after all the additional metal was added in the form of antennas with
grounds  she no longer complained.

I do not claim to understand what goes on around here at an elevation of
about 102 ft. above the valley floor but I believe that we have been
"fortunate"  (or lucky if you prefer that word) and we have so much
metal in
the air that most "discharges" are not of the type that is accompanied
by
thunder!  Two years ago we had a tree die that was no more than 30 ft.
from
the house.  Since I have not always been able to accurately predict
where
trees will fall that I cut myself I hired a professional to take it down
and
he took one look at it, reached up and stripped off a piece of bark and
commented, "See those lines?  This tree was struck by lightning.
That's
what killed it!"  But no lightning related damage to structure or
equipment
here for nearly 15 years!

My only thought on the subject was that only a fool would fail to
disconnect
an antenna that rises above the house roof  during an electrical storm.
I
freely admit that I have several ham radio VHF antennas  (and a GPS
antenna
supplying the disciplined system for frequency reference) well below
roof
peak level  (clear view to south for GPS, of course) that I do not
disconnect from inexpensive radios maintained for the purpose.

This is risk I accept.  And I certainly have seen paths followed by
lightning that are truly amazing!  ....and horrendously damaging.  Never
forget that you only live once and you do have a perfect right to enjoy
your
hobby without fear.

Regards,

Lee A. Mushel  K9WRU

----- Original Message -----
From: "quartz55" quartz55@hughes.net
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Sent: Sunday, September 15, 2013 12:13 PM
Subject: Re: [time-nuts] New NTBW50AA

.  I'm not thrilled having it above the beams, it's the highest metal
thing
in the area and most likely more prone to lightning than anything else
excpet the trees and it will blow around a bit up there as well as
rotating
once in a while.  I did have a strike here once, but it was down by the
barn
and was looking for the underground telephone cables, it blew apart a
post
and knocked off the boards, it was maybe 100 yards from the shack.  It
wouldn't be much of a problem to dig a tiny trench to the fence and I
may
try that after this precise survey is done, then do another and see if
anything changes.  I could hide it i

n a bird box at the fence too.  After all at this point I'm just
playing.
I'll eventually make up my mind.  Also, I don't have an attic, so
that's
not an option.

Dave
N3DT


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"The power of accurate observation is commonly called cynicism by those
who have not got it."
-George Bernard Shaw

Dr. Don Latham AJ7LL
Six Mile Systems LLC
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
Skype: buffler2
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Don, Thanks for the suggestion! I have no doubt that you are correct and especially so at the moment of any "strike" or discharge. But these rods cover an area approximately 200 by 500 ft. and I don't find heavy cable "practical!" Feedline varies from the RG-58 you would expect to 7/8 inch Andrew (350'). Lee ----- Original Message ----- From: "Don Latham" <djl@montana.com> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com> Sent: Sunday, September 15, 2013 3:54 PM Subject: Re: [time-nuts] New NTBW50AA > Lee, if I can make a couple of comments, first, you can have large > voltage differences among your 23 "ground" rods, depending on the ground > terminus of the lightning strokes. Second, antennae and feed lines that > do not attract lightning due to being "hidden" can still be subject to > large induced voltages. > I'd suggest all the ground rods and antenna bases be connected directly > with large cable so they are all truly at the same ground potential... > Don > > Lee Mushel >> Lightning is certainly one of Old Mother Nature's principal weapons as >> well >> as being extremely unpredictable. My House/Shack is on the side of a >> hill >> with the Antenna Farm on top of the hill about 120 yds away with an >> additional 30 ft. of elevation. There is a metal "lightning rod" >> extending >> high enough to provide a "cone of protection" for the house and is >> grounded >> directly. And there are five other metal antennas on that roof, two to >> a >> separate ground and three to still another which is close to the shack >> "ground field" which is three additional ground rods. All ground rods >> I >> mention are the standard 8 ft. copper plated steel items. Each of five >> antennas in the antenna field also have individual ground rods and I >> guess >> if you follow the cable shields you could say that they are all >> connected >> together. A couple of years ago I counted all ground rods and came up >> with 23. Before all of the metal was added and the "lightning rod" was >> the >> only metal extending about 20 ft. above the roof and we had an >> electrical >> storm my wife would complain that she could feel her hair trying to rise >> in >> response to the field coming from the cable which ran across the roof >> and at >> it's closest point was only 6-7 ft. from her head as the crow flies. >> But >> after all the additional metal was added in the form of antennas with >> grounds she no longer complained. >> >> I do not claim to understand what goes on around here at an elevation of >> about 102 ft. above the valley floor but I believe that we have been >> "fortunate" (or lucky if you prefer that word) and we have so much >> metal in >> the air that most "discharges" are not of the type that is accompanied >> by >> thunder! Two years ago we had a tree die that was no more than 30 ft. >> from >> the house. Since I have not always been able to accurately predict >> where >> trees will fall that I cut myself I hired a professional to take it down >> and >> he took one look at it, reached up and stripped off a piece of bark and >> commented, "See those lines? This tree was struck by lightning. >> That's >> what killed it!" But no lightning related damage to structure or >> equipment >> here for nearly 15 years! >> >> My only thought on the subject was that only a fool would fail to >> disconnect >> an antenna that rises above the house roof during an electrical storm. >> I >> freely admit that I have several ham radio VHF antennas (and a GPS >> antenna >> supplying the disciplined system for frequency reference) well below >> roof >> peak level (clear view to south for GPS, of course) that I do not >> disconnect from inexpensive radios maintained for the purpose. >> >> This is risk I accept. And I certainly have seen paths followed by >> lightning that are truly amazing! ....and horrendously damaging. Never >> forget that you only live once and you do have a perfect right to enjoy >> your >> hobby without fear. >> >> Regards, >> >> Lee A. Mushel K9WRU >> >> >> ----- Original Message ----- >> From: "quartz55" <quartz55@hughes.net> >> To: "Discussion of precise time and frequency measurement" >> <time-nuts@febo.com> >> Sent: Sunday, September 15, 2013 12:13 PM >> Subject: Re: [time-nuts] New NTBW50AA >> >> >> . I'm not thrilled having it above the beams, it's the highest metal >> thing >> in the area and most likely more prone to lightning than anything else >> excpet the trees and it will blow around a bit up there as well as >> rotating >> once in a while. I did have a strike here once, but it was down by the >> barn >> and was looking for the underground telephone cables, it blew apart a >> post >> and knocked off the boards, it was maybe 100 yards from the shack. It >> wouldn't be much of a problem to dig a tiny trench to the fence and I >> may >> try that after this precise survey is done, then do another and see if >> anything changes. I could hide it i >>> n a bird box at the fence too. After all at this point I'm just >>> playing. >>> I'll eventually make up my mind. Also, I don't have an attic, so >>> that's >>> not an option. >>> >>> Dave >>> N3DT >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >>> >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > > > -- > "The power of accurate observation is commonly called cynicism by those > who have not got it." > -George Bernard Shaw > > > > Dr. Don Latham AJ7LL > Six Mile Systems LLC > 17850 Six Mile Road > POB 134 > Huson, MT, 59846 > VOX 406-626-4304 > Skype: buffler2 > www.lightningforensics.com > www.sixmilesystems.com > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > >