time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

Re: [time-nuts] PRS 10 and serial port communication

PZ
Philip Zahariev
Fri, Mar 4, 2016 8:36 AM

Hi Jarl,

As I already write PRS10 is configured as serial communication. With oscilloscope I can see serial signals, but when enter this signal in computer with TTL level USB2serial converter all information is repeatable character unknown for me and not as described in manual. For test i change baud rate, number of stop bits, number of bits without success. Also trace signal to U302 MC68HC11 pin 20 and 21 and both TX and RX are the same as on J100 (external connector) - time/period and amplitude has TTL levels. For me looks like internal ROM has incorrect bits or processor is not in normal state and response from serial port coming from different subroutine.
I expect hints how to check processor state. Or how to push processor in state where will have normal serial communication.

Best regards
Philip

On 4.03.2016 г., at 9:35, Jarl Risum jarl.risum@gmail.com wrote:

OK, Philip

If you have confirmed with a Oscilloscope that no RS-232 communication is present on the TXD output pin, I sugggest you check the status of R247/R248 and R353/R354 - see the manuals chapter on "PRS 10 Applications" (page 20), description of pin 4 and pin 7.

Good luck

Jarl

2016-03-03 20:08 GMT+01:00 Philip Zahariev philip_zahariev@yahoo.com:

Hi Jarl,

Of course I use TTL level USB serial converter. Unfortunately, as I write no communication, or at least not what I expect.

Best regards

Philip

From: Jarl Risum [mailto:jarl.risum@gmail.com]
Sent: Wednesday, March 2, 2016 11:49 PM
To: philip.zahariev@yahoo.com; philip_zahariev@yahoo.com
Subject: PRS 10 and serial port communication

Philip,

The outgoing RS-232 port from PRS 10 is a TTL inverter with an approx. 1 kohm series resistor.

See diagram of the microcontroller board, upper right corner.

For this reason you may have problems reading the outgoing RS-232 messages with a standard RS-232 PC input. But you may see the transmission taking place with a oscilloscope.

I have had the same problem with my PRS 10 and solved it with a simple 2 transistor non-inverting amplifier fed from a +12V source.

If interested, I can send you the diagram.

Cheers from

OZ9MO / Jarl

Hi Jarl, As I already write PRS10 is configured as serial communication. With oscilloscope I can see serial signals, but when enter this signal in computer with TTL level USB2serial converter all information is repeatable character unknown for me and not as described in manual. For test i change baud rate, number of stop bits, number of bits without success. Also trace signal to U302 MC68HC11 pin 20 and 21 and both TX and RX are the same as on J100 (external connector) - time/period and amplitude has TTL levels. For me looks like internal ROM has incorrect bits or processor is not in normal state and response from serial port coming from different subroutine. I expect hints how to check processor state. Or how to push processor in state where will have normal serial communication. Best regards Philip > On 4.03.2016 г., at 9:35, Jarl Risum <jarl.risum@gmail.com> wrote: > > OK, Philip > > If you have confirmed with a Oscilloscope that no RS-232 communication is present on the TXD output pin, I sugggest you check the status of R247/R248 and R353/R354 - see the manuals chapter on "PRS 10 Applications" (page 20), description of pin 4 and pin 7. > > Good luck > > Jarl > > 2016-03-03 20:08 GMT+01:00 Philip Zahariev <philip_zahariev@yahoo.com>: >> Hi Jarl, >> >> >> >> Of course I use TTL level USB serial converter. Unfortunately, as I write no communication, or at least not what I expect. >> >> >> >> Best regards >> >> Philip >> >> >> >> From: Jarl Risum [mailto:jarl.risum@gmail.com] >> Sent: Wednesday, March 2, 2016 11:49 PM >> To: philip.zahariev@yahoo.com; philip_zahariev@yahoo.com >> Subject: PRS 10 and serial port communication >> >> >> >> Philip, >> >> The outgoing RS-232 port from PRS 10 is a TTL inverter with an approx. 1 kohm series resistor. >> >> See diagram of the microcontroller board, upper right corner. >> >> For this reason you may have problems reading the outgoing RS-232 messages with a standard RS-232 PC input. But you may see the transmission taking place with a oscilloscope. >> >> I have had the same problem with my PRS 10 and solved it with a simple 2 transistor non-inverting amplifier fed from a +12V source. >> >> If interested, I can send you the diagram. >> >> Cheers from >> >> OZ9MO / Jarl >> >
MC
Mike Cook
Fri, Mar 4, 2016 9:34 AM

You could try « RC 1 » blind.
That will reset the EEPROM values to factory settings. However I do not see any settings for serial rates, so it may not help.
From your oscilloscope, it should be possible to determine stop/start and baud rate and decode the binary data.
I have done this in the distant past.
There are web resources. Check out http://electronics.stackexchange.com/questions/17562/how-to-read-serial-data-from-oscilloscope
for example.

Mike

Le 4 mars 2016 à 09:36, Philip Zahariev via time-nuts time-nuts@febo.com a écrit :

Hi Jarl,

As I already write PRS10 is configured as serial communication. With oscilloscope I can see serial signals, but when enter this signal in computer with TTL level USB2serial converter all information is repeatable character unknown for me and not as described in manual. For test i change baud rate, number of stop bits, number of bits without success. Also trace signal to U302 MC68HC11 pin 20 and 21 and both TX and RX are the same as on J100 (external connector) - time/period and amplitude has TTL levels. For me looks like internal ROM has incorrect bits or processor is not in normal state and response from serial port coming from different subroutine.
I expect hints how to check processor state. Or how to push processor in state where will have normal serial communication.

Best regards
Philip

On 4.03.2016 г., at 9:35, Jarl Risum jarl.risum@gmail.com wrote:

OK, Philip

If you have confirmed with a Oscilloscope that no RS-232 communication is present on the TXD output pin, I sugggest you check the status of R247/R248 and R353/R354 - see the manuals chapter on "PRS 10 Applications" (page 20), description of pin 4 and pin 7.

Good luck

Jarl

2016-03-03 20:08 GMT+01:00 Philip Zahariev philip_zahariev@yahoo.com:

Hi Jarl,

Of course I use TTL level USB serial converter. Unfortunately, as I write no communication, or at least not what I expect.

Best regards

Philip

From: Jarl Risum [mailto:jarl.risum@gmail.com]
Sent: Wednesday, March 2, 2016 11:49 PM
To: philip.zahariev@yahoo.com; philip_zahariev@yahoo.com
Subject: PRS 10 and serial port communication

Philip,

The outgoing RS-232 port from PRS 10 is a TTL inverter with an approx. 1 kohm series resistor.

See diagram of the microcontroller board, upper right corner.

For this reason you may have problems reading the outgoing RS-232 messages with a standard RS-232 PC input. But you may see the transmission taking place with a oscilloscope.

I have had the same problem with my PRS 10 and solved it with a simple 2 transistor non-inverting amplifier fed from a +12V source.

If interested, I can send you the diagram.

Cheers from

OZ9MO / Jarl


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

"The main function of a modern police force is filling in forms."

You could try « RC 1 » blind. That will reset the EEPROM values to factory settings. However I do not see any settings for serial rates, so it may not help. From your oscilloscope, it should be possible to determine stop/start and baud rate and decode the binary data. I have done this in the distant past. There are web resources. Check out <http://electronics.stackexchange.com/questions/17562/how-to-read-serial-data-from-oscilloscope> for example. Mike > Le 4 mars 2016 à 09:36, Philip Zahariev via time-nuts <time-nuts@febo.com> a écrit : > > Hi Jarl, > > As I already write PRS10 is configured as serial communication. With oscilloscope I can see serial signals, but when enter this signal in computer with TTL level USB2serial converter all information is repeatable character unknown for me and not as described in manual. For test i change baud rate, number of stop bits, number of bits without success. Also trace signal to U302 MC68HC11 pin 20 and 21 and both TX and RX are the same as on J100 (external connector) - time/period and amplitude has TTL levels. For me looks like internal ROM has incorrect bits or processor is not in normal state and response from serial port coming from different subroutine. > I expect hints how to check processor state. Or how to push processor in state where will have normal serial communication. > > Best regards > Philip > >> On 4.03.2016 г., at 9:35, Jarl Risum <jarl.risum@gmail.com> wrote: >> >> OK, Philip >> >> If you have confirmed with a Oscilloscope that no RS-232 communication is present on the TXD output pin, I sugggest you check the status of R247/R248 and R353/R354 - see the manuals chapter on "PRS 10 Applications" (page 20), description of pin 4 and pin 7. >> >> Good luck >> >> Jarl >> >> 2016-03-03 20:08 GMT+01:00 Philip Zahariev <philip_zahariev@yahoo.com>: >>> Hi Jarl, >>> >>> >>> >>> Of course I use TTL level USB serial converter. Unfortunately, as I write no communication, or at least not what I expect. >>> >>> >>> >>> Best regards >>> >>> Philip >>> >>> >>> >>> From: Jarl Risum [mailto:jarl.risum@gmail.com] >>> Sent: Wednesday, March 2, 2016 11:49 PM >>> To: philip.zahariev@yahoo.com; philip_zahariev@yahoo.com >>> Subject: PRS 10 and serial port communication >>> >>> >>> >>> Philip, >>> >>> The outgoing RS-232 port from PRS 10 is a TTL inverter with an approx. 1 kohm series resistor. >>> >>> See diagram of the microcontroller board, upper right corner. >>> >>> For this reason you may have problems reading the outgoing RS-232 messages with a standard RS-232 PC input. But you may see the transmission taking place with a oscilloscope. >>> >>> I have had the same problem with my PRS 10 and solved it with a simple 2 transistor non-inverting amplifier fed from a +12V source. >>> >>> If interested, I can send you the diagram. >>> >>> Cheers from >>> >>> OZ9MO / Jarl >>> >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. "The main function of a modern police force is filling in forms."
PZ
Philip Zahariev
Tue, Mar 8, 2016 12:56 AM

Hello group,

I have significant progress with generator. And there were new questions :-)
Frequency was little bit higher than 10 MHz - after correction of PLL constants SP and MO now is just on spot.

In documentation is not clear described what is active state of signal 1PPS_IN. I found that signal must be 1 and goes to 0 during PPS pulse. After experience with SERIAL line inverter this time was easy (and have free inverters). Opposite is situation of output 1PPS_OUTPUT - his state is 0 and goes 1 during PPS pulse :-)

  1. I found that AD8 - Amplified ac photosignal is changing between 1.120 to 3.991 and any value are different than previous, sometimes jumps are more than 2.xxx (monitoring with RBMon) - Is this usual?

  2. AD10 - Case temperature (10 mV/°C) has value 0.781 which means 78°C. I remember that temperature has to be below 40°C, but when mount 2 heatsinks from computer CPU (socket 370) temperature decrease to 67-68°C. Do I need to install fans (don't like noise) or increase size? Are this 40°C was measured with external thermometer or means that AD10 need to be lower than 0.4xx?

  3. Firmware of my device is 3.15. Before value for TO - Time-tag offset (1816) from RBMon has Red square. According manual in this version of firmware TO is not user changeable. Is it available file with new firmware version? Processor on device is MC68HC711E20CFN3 - that means I need to buy new empty processor and program it. What improvements was made in newer version of firmware?

Best regards
Philip

Hello group, I have significant progress with generator. And there were new questions :-) Frequency was little bit higher than 10 MHz - after correction of PLL constants SP and MO now is just on spot. In documentation is not clear described what is active state of signal 1PPS_IN. I found that signal must be 1 and goes to 0 during PPS pulse. After experience with SERIAL line inverter this time was easy (and have free inverters). Opposite is situation of output 1PPS_OUTPUT - his state is 0 and goes 1 during PPS pulse :-) 1. I found that AD8 - Amplified ac photosignal is changing between 1.120 to 3.991 and any value are different than previous, sometimes jumps are more than 2.xxx (monitoring with RBMon) - Is this usual? 2. AD10 - Case temperature (10 mV/°C) has value 0.781 which means 78°C. I remember that temperature has to be below 40°C, but when mount 2 heatsinks from computer CPU (socket 370) temperature decrease to 67-68°C. Do I need to install fans (don't like noise) or increase size? Are this 40°C was measured with external thermometer or means that AD10 need to be lower than 0.4xx? 3. Firmware of my device is 3.15. Before value for TO - Time-tag offset (1816) from RBMon has Red square. According manual in this version of firmware TO is not user changeable. Is it available file with new firmware version? Processor on device is MC68HC711E20CFN3 - that means I need to buy new empty processor and program it. What improvements was made in newer version of firmware? Best regards Philip
MC
Mike Cook
Tue, Mar 8, 2016 7:36 AM

Le 8 mars 2016 à 01:56, Philip Zahariev via time-nuts time-nuts@febo.com a écrit :

Hello group,

I have significant progress with generator. And there were new questions :-)
Frequency was little bit higher than 10 MHz - after correction of PLL constants SP and MO now is just on spot.

In documentation is not clear described what is active state of signal 1PPS_IN. I found that signal must be 1 and goes to 0 during PPS pulse. After experience with SERIAL line inverter this time was easy (and have free inverters). Opposite is situation of output 1PPS_OUTPUT - his state is 0 and goes 1 during PPS pulse :-)

  1. I found that AD8 - Amplified ac photosignal is changing between 1.120 to 3.991 and any value are different than previous, sometimes jumps are more than 2.xxx (monitoring with RBMon) - Is this usual?

I think so. At least I have that order of variability on mine.

  1. AD10 - Case temperature (10 mV/°C) has value 0.781 which means 78°C. I remember that temperature has to be below 40°C, but when mount 2 heatsinks from computer CPU (socket 370) temperature decrease to 67-68°C. Do I need to install fans (don't like noise) or increase size? Are this 40°C was measured with external thermometer or means that AD10 need to be lower than 0.4xx?

My PRS10s are in the same ball park (0.736) on one I am looking at. I have installed SRS supplied Bench Heat Sinks on them and the measured external baseplate temperature is 42.9°. If it’s good enough for SRS then it’s good enough for me though I suspect that within reason lower is better. However, that said, I don’t think that SRS put a fan in the 725 which is a PRS10 + buffered outputs and that is in a case where temps are likely to be higher..

  1. Firmware of my device is 3.15. Before value for TO - Time-tag offset (1816) from RBMon has Red square. According manual in this version of firmware TO is not user changeable. Is it available file with new firmware version? Processor on device is MC68HC711E20CFN3 - that means I need to buy new empty processor and program it. What improvements was made in newer version of firmware?

That is not strictly true. With 3.15, you should be able to change the current value but you cannot write it to EEPROM (TO!) so that the new value is used at power on. With later versions ex. 3.24 you can change the EEPROM value.

I am not sure why it is flagged RED as from my experience the colors used are.
None:  in spec or not different from EEPROM
Black:  different from EEPROM value
Blue:    nearing a limit.
Red:    Out of spec (ex. SF hit the wall at -2000)

Equally the black flag on your FC values is strange as the stats value is in limits. If you use FC!? to see what the EEPROM values are you will see 4 values, including the stats figure. From the one in front of me:
47 number of power cycles
5168 number of times DAC values written to EEPROM ( twice a day from my experience so you can get a rough idea of accumulated power on time)
1532 high and
2056 low DAC values used at reset/power on

Best regards
Philip


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

"The main function of a modern police force is filling in forms."

> Le 8 mars 2016 à 01:56, Philip Zahariev via time-nuts <time-nuts@febo.com> a écrit : > > Hello group, > > I have significant progress with generator. And there were new questions :-) > Frequency was little bit higher than 10 MHz - after correction of PLL constants SP and MO now is just on spot. > > In documentation is not clear described what is active state of signal 1PPS_IN. I found that signal must be 1 and goes to 0 during PPS pulse. After experience with SERIAL line inverter this time was easy (and have free inverters). Opposite is situation of output 1PPS_OUTPUT - his state is 0 and goes 1 during PPS pulse :-) > > 1. I found that AD8 - Amplified ac photosignal is changing between 1.120 to 3.991 and any value are different than previous, sometimes jumps are more than 2.xxx (monitoring with RBMon) - Is this usual? I think so. At least I have that order of variability on mine. > > 2. AD10 - Case temperature (10 mV/°C) has value 0.781 which means 78°C. I remember that temperature has to be below 40°C, but when mount 2 heatsinks from computer CPU (socket 370) temperature decrease to 67-68°C. Do I need to install fans (don't like noise) or increase size? Are this 40°C was measured with external thermometer or means that AD10 need to be lower than 0.4xx? > My PRS10s are in the same ball park (0.736) on one I am looking at. I have installed SRS supplied Bench Heat Sinks on them and the measured external baseplate temperature is 42.9°. If it’s good enough for SRS then it’s good enough for me though I suspect that within reason lower is better. However, that said, I don’t think that SRS put a fan in the 725 which is a PRS10 + buffered outputs and that is in a case where temps are likely to be higher.. > 3. Firmware of my device is 3.15. Before value for TO - Time-tag offset (1816) from RBMon has Red square. According manual in this version of firmware TO is not user changeable. Is it available file with new firmware version? Processor on device is MC68HC711E20CFN3 - that means I need to buy new empty processor and program it. What improvements was made in newer version of firmware? > That is not strictly true. With 3.15, you should be able to change the current value but you cannot write it to EEPROM (TO!) so that the new value is used at power on. With later versions ex. 3.24 you can change the EEPROM value. I am not sure why it is flagged RED as from my experience the colors used are. None: in spec or not different from EEPROM Black: different from EEPROM value Blue: nearing a limit. Red: Out of spec (ex. SF hit the wall at -2000) Equally the black flag on your FC values is strange as the stats value is in limits. If you use FC!? to see what the EEPROM values are you will see 4 values, including the stats figure. From the one in front of me: 47 number of power cycles 5168 number of times DAC values written to EEPROM ( twice a day from my experience so you can get a rough idea of accumulated power on time) 1532 high and 2056 low DAC values used at reset/power on > Best regards > Philip > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. "The main function of a modern police force is filling in forms."
BC
Bob Camp
Tue, Mar 8, 2016 11:57 AM

Hi

I seem to have a lot of experience running Rb’s without heatsinks. No matter how many chickens I sacrifice at the full moon, the results seem to always come out the same. If you don’t get the temperature of the internal baseplate down under 40C (with a heatsink and maybe fan) they die. These little guys all get very hot. They have to dump heat to keep the offset between the two sections of the physics package at the right offset. No matter what the brand, they run for a year or four and croak. The average is about two years.

One would think that I’d stop this experiment ….unfortunately, I have help from others who also seem to believe in the power of the chickens as opposed to heatsinks.

Bob

On Mar 8, 2016, at 2:36 AM, Mike Cook michael.cook@sfr.fr wrote:

Le 8 mars 2016 à 01:56, Philip Zahariev via time-nuts time-nuts@febo.com a écrit :

Hello group,

I have significant progress with generator. And there were new questions :-)
Frequency was little bit higher than 10 MHz - after correction of PLL constants SP and MO now is just on spot.

In documentation is not clear described what is active state of signal 1PPS_IN. I found that signal must be 1 and goes to 0 during PPS pulse. After experience with SERIAL line inverter this time was easy (and have free inverters). Opposite is situation of output 1PPS_OUTPUT - his state is 0 and goes 1 during PPS pulse :-)

  1. I found that AD8 - Amplified ac photosignal is changing between 1.120 to 3.991 and any value are different than previous, sometimes jumps are more than 2.xxx (monitoring with RBMon) - Is this usual?

I think so. At least I have that order of variability on mine.

  1. AD10 - Case temperature (10 mV/°C) has value 0.781 which means 78°C. I remember that temperature has to be below 40°C, but when mount 2 heatsinks from computer CPU (socket 370) temperature decrease to 67-68°C. Do I need to install fans (don't like noise) or increase size? Are this 40°C was measured with external thermometer or means that AD10 need to be lower than 0.4xx?

My PRS10s are in the same ball park (0.736) on one I am looking at. I have installed SRS supplied Bench Heat Sinks on them and the measured external baseplate temperature is 42.9°. If it’s good enough for SRS then it’s good enough for me though I suspect that within reason lower is better. However, that said, I don’t think that SRS put a fan in the 725 which is a PRS10 + buffered outputs and that is in a case where temps are likely to be higher..

  1. Firmware of my device is 3.15. Before value for TO - Time-tag offset (1816) from RBMon has Red square. According manual in this version of firmware TO is not user changeable. Is it available file with new firmware version? Processor on device is MC68HC711E20CFN3 - that means I need to buy new empty processor and program it. What improvements was made in newer version of firmware?

That is not strictly true. With 3.15, you should be able to change the current value but you cannot write it to EEPROM (TO!) so that the new value is used at power on. With later versions ex. 3.24 you can change the EEPROM value.

I am not sure why it is flagged RED as from my experience the colors used are.
None:  in spec or not different from EEPROM
Black:  different from EEPROM value
Blue:    nearing a limit.
Red:    Out of spec (ex. SF hit the wall at -2000)

Equally the black flag on your FC values is strange as the stats value is in limits. If you use FC!? to see what the EEPROM values are you will see 4 values, including the stats figure. From the one in front of me:
47 number of power cycles
5168 number of times DAC values written to EEPROM ( twice a day from my experience so you can get a rough idea of accumulated power on time)
1532 high and
2056 low DAC values used at reset/power on

Best regards
Philip


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

"The main function of a modern police force is filling in forms."


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi I seem to have a lot of experience running Rb’s without heatsinks. No matter how many chickens I sacrifice at the full moon, the results seem to always come out the same. If you don’t get the temperature of the internal baseplate down under 40C (with a heatsink and maybe fan) they die. These little guys all get very hot. They have to dump heat to keep the offset between the two sections of the physics package at the right offset. No matter what the brand, they run for a year or four and croak. The average is about two years. One *would* think that I’d stop this experiment ….unfortunately, I have help from others who also seem to believe in the power of the chickens as opposed to heatsinks. Bob > On Mar 8, 2016, at 2:36 AM, Mike Cook <michael.cook@sfr.fr> wrote: > > >> Le 8 mars 2016 à 01:56, Philip Zahariev via time-nuts <time-nuts@febo.com> a écrit : >> >> Hello group, >> >> I have significant progress with generator. And there were new questions :-) >> Frequency was little bit higher than 10 MHz - after correction of PLL constants SP and MO now is just on spot. >> >> In documentation is not clear described what is active state of signal 1PPS_IN. I found that signal must be 1 and goes to 0 during PPS pulse. After experience with SERIAL line inverter this time was easy (and have free inverters). Opposite is situation of output 1PPS_OUTPUT - his state is 0 and goes 1 during PPS pulse :-) >> >> 1. I found that AD8 - Amplified ac photosignal is changing between 1.120 to 3.991 and any value are different than previous, sometimes jumps are more than 2.xxx (monitoring with RBMon) - Is this usual? > > I think so. At least I have that order of variability on mine. >> >> 2. AD10 - Case temperature (10 mV/°C) has value 0.781 which means 78°C. I remember that temperature has to be below 40°C, but when mount 2 heatsinks from computer CPU (socket 370) temperature decrease to 67-68°C. Do I need to install fans (don't like noise) or increase size? Are this 40°C was measured with external thermometer or means that AD10 need to be lower than 0.4xx? >> > > My PRS10s are in the same ball park (0.736) on one I am looking at. I have installed SRS supplied Bench Heat Sinks on them and the measured external baseplate temperature is 42.9°. If it’s good enough for SRS then it’s good enough for me though I suspect that within reason lower is better. However, that said, I don’t think that SRS put a fan in the 725 which is a PRS10 + buffered outputs and that is in a case where temps are likely to be higher.. > >> 3. Firmware of my device is 3.15. Before value for TO - Time-tag offset (1816) from RBMon has Red square. According manual in this version of firmware TO is not user changeable. Is it available file with new firmware version? Processor on device is MC68HC711E20CFN3 - that means I need to buy new empty processor and program it. What improvements was made in newer version of firmware? >> > > That is not strictly true. With 3.15, you should be able to change the current value but you cannot write it to EEPROM (TO!) so that the new value is used at power on. With later versions ex. 3.24 you can change the EEPROM value. > > I am not sure why it is flagged RED as from my experience the colors used are. > None: in spec or not different from EEPROM > Black: different from EEPROM value > Blue: nearing a limit. > Red: Out of spec (ex. SF hit the wall at -2000) > > Equally the black flag on your FC values is strange as the stats value is in limits. If you use FC!? to see what the EEPROM values are you will see 4 values, including the stats figure. From the one in front of me: > 47 number of power cycles > 5168 number of times DAC values written to EEPROM ( twice a day from my experience so you can get a rough idea of accumulated power on time) > 1532 high and > 2056 low DAC values used at reset/power on > >> Best regards >> Philip >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > "The main function of a modern police force is filling in forms." > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.