Peter Reilley
Tue, Oct 18, 2016 1:11 PM

I bought an Isotemp OCXO82-59 with a frequency of 10 MHz for a $3 at the
MIT flea market.
As expected it was dead.  It heats up as expected but looking at the
output with a scope there
is nothing.  However looking at the output with a spectrum analyzer I
can see a faint 10 MHz
signal.  It seems that the oscillator is running but the output
circuitry is dead.  Reasonable
assumption?

Anyway, has anyone had any luck unsoldering the tin case without
destroying it?

Pete.

I bought an Isotemp OCXO82-59 with a frequency of 10 MHz for a $3 at the MIT flea market. As expected it was dead. It heats up as expected but looking at the output with a scope there is nothing. However looking at the output with a spectrum analyzer I can see a faint 10 MHz signal. It seems that the oscillator is running but the output circuitry is dead. Reasonable assumption? Anyway, has anyone had any luck unsoldering the tin case without destroying it? Pete.
PS
paul swed
Tue, Oct 18, 2016 2:14 PM

Pete
Another MIT-er. Well the only way I have handled them is with a small torch.
Clamp the unit. Heat the solder and it will flow down to the low point and
tend to drip out.
Stick a small xacto knife in to break any remaining solder loose.
Not super pretty but all of it can be cleaned up usually and then
re-soldered.
Hey its dead whats to loose?
Regards
Paul
WB8TSL

On Tue, Oct 18, 2016 at 9:11 AM, Peter Reilley preilley_454@comcast.net
wrote:

I bought an Isotemp OCXO82-59 with a frequency of 10 MHz for a $3 at the
MIT flea market.
As expected it was dead.  It heats up as expected but looking at the
output with a scope there
is nothing.  However looking at the output with a spectrum analyzer I can
see a faint 10 MHz
signal.  It seems that the oscillator is running but the output circuitry
is dead.  Reasonable
assumption?

Anyway, has anyone had any luck unsoldering the tin case without
destroying it?

Pete.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

Pete Another MIT-er. Well the only way I have handled them is with a small torch. Clamp the unit. Heat the solder and it will flow down to the low point and tend to drip out. Stick a small xacto knife in to break any remaining solder loose. Not super pretty but all of it can be cleaned up usually and then re-soldered. Hey its dead whats to loose? Regards Paul WB8TSL On Tue, Oct 18, 2016 at 9:11 AM, Peter Reilley <preilley_454@comcast.net> wrote: > I bought an Isotemp OCXO82-59 with a frequency of 10 MHz for a $3 at the > MIT flea market. > As expected it was dead. It heats up as expected but looking at the > output with a scope there > is nothing. However looking at the output with a spectrum analyzer I can > see a faint 10 MHz > signal. It seems that the oscillator is running but the output circuitry > is dead. Reasonable > assumption? > > Anyway, has anyone had any luck unsoldering the tin case without > destroying it? > > Pete. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/m > ailman/listinfo/time-nuts > and follow the instructions there. >
JL
J. L. Trantham
Tue, Oct 18, 2016 2:57 PM

Pete,

I'm not familiar with your OCXO but I found one shown on 'theBay' (item  261920574725) and it appeared to have an option for 'mounting screws', four of them, on the bottom.  Interestingly, the 'link' to the datasheet for that unit did not show threads for mounting screws.

If your unit has that option, I would suggest placing four long screws, mounting the item in a vise, use a small torch (I've used a hand held propane torch turned down very low to open a number of units from 5061A's) around the bottom of the case while gripping the top with an appropriate sized Channel Lock plier and lifting off the top.

If you can repair the OCXO, it should be easy to reassemble the unit with solder.

TheBay unit looks like it has a screw cover (which likely has a rubber gasket) for mechanical adjustment of the frequency.  I'd remove that before applying the torch. :^).

If you get it open, I'd love to see some pictures of the insides.

Good luck and hope this helps.

Joe

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Peter Reilley
Sent: Tuesday, October 18, 2016 8:11 AM
To: time-nuts@febo.com
Subject: [time-nuts] Opening an Isotemp OCXO

I bought an Isotemp OCXO82-59 with a frequency of 10 MHz for a $3 at the MIT flea market.
As expected it was dead.  It heats up as expected but looking at the
output with a scope there
is nothing.  However looking at the output with a spectrum analyzer I
can see a faint 10 MHz
signal.  It seems that the oscillator is running but the output
circuitry is dead.  Reasonable
assumption?

Anyway, has anyone had any luck unsoldering the tin case without destroying it?

Pete.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Pete, I'm not familiar with your OCXO but I found one shown on 'theBay' (item 261920574725) and it appeared to have an option for 'mounting screws', four of them, on the bottom. Interestingly, the 'link' to the datasheet for that unit did not show threads for mounting screws. If your unit has that option, I would suggest placing four long screws, mounting the item in a vise, use a small torch (I've used a hand held propane torch turned down very low to open a number of units from 5061A's) around the bottom of the case while gripping the top with an appropriate sized Channel Lock plier and lifting off the top. If you can repair the OCXO, it should be easy to reassemble the unit with solder. TheBay unit looks like it has a screw cover (which likely has a rubber gasket) for mechanical adjustment of the frequency. I'd remove that before applying the torch. :^). If you get it open, I'd love to see some pictures of the insides. Good luck and hope this helps. Joe -----Original Message----- From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Peter Reilley Sent: Tuesday, October 18, 2016 8:11 AM To: time-nuts@febo.com Subject: [time-nuts] Opening an Isotemp OCXO I bought an Isotemp OCXO82-59 with a frequency of 10 MHz for a $3 at the MIT flea market. As expected it was dead. It heats up as expected but looking at the output with a scope there is nothing. However looking at the output with a spectrum analyzer I can see a faint 10 MHz signal. It seems that the oscillator is running but the output circuitry is dead. Reasonable assumption? Anyway, has anyone had any luck unsoldering the tin case without destroying it? Pete. _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
CH
Christopher Hoover
Tue, Oct 18, 2016 4:26 PM

I got one open without too much much trouble with a propane torch.  Work
fast, use gravity and have something pointy to pry with as you go.

-ch
73 de ai6kg

On Oct 18, 2016 6:14 AM, "Peter Reilley" preilley_454@comcast.net wrote:

I bought an Isotemp OCXO82-59 with a frequency of 10 MHz for a $3 at the
MIT flea market.
As expected it was dead.  It heats up as expected but looking at the
output with a scope there
is nothing.  However looking at the output with a spectrum analyzer I can
see a faint 10 MHz
signal.  It seems that the oscillator is running but the output circuitry
is dead.  Reasonable
assumption?

Anyway, has anyone had any luck unsoldering the tin case without
destroying it?

Pete.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

I got one open without too much much trouble with a propane torch. Work fast, use gravity and have something pointy to pry with as you go. -ch 73 de ai6kg On Oct 18, 2016 6:14 AM, "Peter Reilley" <preilley_454@comcast.net> wrote: > I bought an Isotemp OCXO82-59 with a frequency of 10 MHz for a $3 at the > MIT flea market. > As expected it was dead. It heats up as expected but looking at the > output with a scope there > is nothing. However looking at the output with a spectrum analyzer I can > see a faint 10 MHz > signal. It seems that the oscillator is running but the output circuitry > is dead. Reasonable > assumption? > > Anyway, has anyone had any luck unsoldering the tin case without > destroying it? > > Pete. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/m > ailman/listinfo/time-nuts > and follow the instructions there. >
PR
Peter Reilley
Fri, Oct 28, 2016 3:31 PM

I did finally get it open.  I used a very large old style soldering
iron and .003 inch steel shim stock.  I would melt the solder on the
straight seams and insert small pieces of the shim.  Solder does not
stick well to steel so the shim kept the soldered seam open.
I used a soldering iron rather than a torch because I can control the
temperature.

I could not use the shim at the corners.  After all the straight seams
were separated I could pull each corner using a screw in the mounting
hole and melt the solder at the corner.  Slowly working my way around,
corner by corner, I got it opened.  I did not damage anything so I
should be able to close it up after I fix it.

Looking around with my scope it seems that the output driver chip is bad
as I expected.  It is a TI 14 pin surface mount DIP.  It says S30 on it
which if it is a 74S30 it is an 8 input positive NAND gate.  The board
layout confirms this as the 10 MHz signal is connected to pin 2 and all
other inputs are tied high.  Pin 8 is connected to the output.

The chip is run off 12 volts so it must be CMOS.  But I cannot find any
chip like that that will run off 12 volts.  Any suggestions for a replacement?

Also, using an 8 input NAND chip for a driver seems an odd choice.

When I put 12 volts on the unit the S30 chip gets really hot. After I
removed the chip the unit seems to work OK.  The current jumps between
about .1 amp to .9 amps.  It seems like the temperature regulator is
an on/off type controller.

The device on eBay, item 261920574725, looks exactly like what I have.

I have placed a bunch of pictures in my dropbox.
https://www.dropbox.com/sh/52e9d1rva9kpb3w/AABmbIj1aK7Zk2J9SNMmu-JAa?dl=0

Pete.

On 10/18/2016 10:57 AM, J. L. Trantham wrote:

Pete,

I'm not familiar with your OCXO but I found one shown on 'theBay' (item  261920574725) and it appeared to have an option for 'mounting screws', four of them, on the bottom.  Interestingly, the 'link' to the datasheet for that unit did not show threads for mounting screws.

If your unit has that option, I would suggest placing four long screws, mounting the item in a vise, use a small torch (I've used a hand held propane torch turned down very low to open a number of units from 5061A's) around the bottom of the case while gripping the top with an appropriate sized Channel Lock plier and lifting off the top.

If you can repair the OCXO, it should be easy to reassemble the unit with solder.

TheBay unit looks like it has a screw cover (which likely has a rubber gasket) for mechanical adjustment of the frequency.  I'd remove that before applying the torch. :^).

If you get it open, I'd love to see some pictures of the insides.

Good luck and hope this helps.

Joe

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Peter Reilley
Sent: Tuesday, October 18, 2016 8:11 AM
To: time-nuts@febo.com
Subject: [time-nuts] Opening an Isotemp OCXO

I bought an Isotemp OCXO82-59 with a frequency of 10 MHz for a $3 at the MIT flea market.
As expected it was dead.  It heats up as expected but looking at the
output with a scope there
is nothing.  However looking at the output with a spectrum analyzer I
can see a faint 10 MHz
signal.  It seems that the oscillator is running but the output
circuitry is dead.  Reasonable
assumption?

Anyway, has anyone had any luck unsoldering the tin case without destroying it?

Pete.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

I did finally get it open. I used a very large old style soldering iron and .003 inch steel shim stock. I would melt the solder on the straight seams and insert small pieces of the shim. Solder does not stick well to steel so the shim kept the soldered seam open. I used a soldering iron rather than a torch because I can control the temperature. I could not use the shim at the corners. After all the straight seams were separated I could pull each corner using a screw in the mounting hole and melt the solder at the corner. Slowly working my way around, corner by corner, I got it opened. I did not damage anything so I should be able to close it up after I fix it. Looking around with my scope it seems that the output driver chip is bad as I expected. It is a TI 14 pin surface mount DIP. It says S30 on it which if it is a 74S30 it is an 8 input positive NAND gate. The board layout confirms this as the 10 MHz signal is connected to pin 2 and all other inputs are tied high. Pin 8 is connected to the output. The chip is run off 12 volts so it must be CMOS. But I cannot find any chip like that that will run off 12 volts. Any suggestions for a replacement? Also, using an 8 input NAND chip for a driver seems an odd choice. When I put 12 volts on the unit the S30 chip gets really hot. After I removed the chip the unit seems to work OK. The current jumps between about .1 amp to .9 amps. It seems like the temperature regulator is an on/off type controller. The device on eBay, item 261920574725, looks exactly like what I have. I have placed a bunch of pictures in my dropbox. https://www.dropbox.com/sh/52e9d1rva9kpb3w/AABmbIj1aK7Zk2J9SNMmu-JAa?dl=0 Pete. On 10/18/2016 10:57 AM, J. L. Trantham wrote: > Pete, > > I'm not familiar with your OCXO but I found one shown on 'theBay' (item 261920574725) and it appeared to have an option for 'mounting screws', four of them, on the bottom. Interestingly, the 'link' to the datasheet for that unit did not show threads for mounting screws. > > If your unit has that option, I would suggest placing four long screws, mounting the item in a vise, use a small torch (I've used a hand held propane torch turned down very low to open a number of units from 5061A's) around the bottom of the case while gripping the top with an appropriate sized Channel Lock plier and lifting off the top. > > If you can repair the OCXO, it should be easy to reassemble the unit with solder. > > TheBay unit looks like it has a screw cover (which likely has a rubber gasket) for mechanical adjustment of the frequency. I'd remove that before applying the torch. :^). > > If you get it open, I'd love to see some pictures of the insides. > > Good luck and hope this helps. > > Joe > > > > -----Original Message----- > From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Peter Reilley > Sent: Tuesday, October 18, 2016 8:11 AM > To: time-nuts@febo.com > Subject: [time-nuts] Opening an Isotemp OCXO > > I bought an Isotemp OCXO82-59 with a frequency of 10 MHz for a $3 at the MIT flea market. > As expected it was dead. It heats up as expected but looking at the > output with a scope there > is nothing. However looking at the output with a spectrum analyzer I > can see a faint 10 MHz > signal. It seems that the oscillator is running but the output > circuitry is dead. Reasonable > assumption? > > Anyway, has anyone had any luck unsoldering the tin case without destroying it? > > Pete. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > >
PK
Poul-Henning Kamp
Fri, Oct 28, 2016 3:49 PM

In message 10a3ea7d-37f0-51bc-2470-35645d767bbd@comcast.net, Peter Reilley writes:

The chip is run off 12 volts so it must be CMOS.

Or the OCXO is not a 12V model ?

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

-------- In message <10a3ea7d-37f0-51bc-2470-35645d767bbd@comcast.net>, Peter Reilley writes: >The chip is run off 12 volts so it must be CMOS. Or the OCXO is not a 12V model ? -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
PR
Peter Reilley
Fri, Oct 28, 2016 4:09 PM

The only document that I could find said 12 volt.

Pete.

On 10/28/2016 11:49 AM, Poul-Henning Kamp wrote:


In message 10a3ea7d-37f0-51bc-2470-35645d767bbd@comcast.net, Peter Reilley writes:

The chip is run off 12 volts so it must be CMOS.

Or the OCXO is not a 12V model ?

The only document that I could find said 12 volt. Pete. On 10/28/2016 11:49 AM, Poul-Henning Kamp wrote: > -------- > In message <10a3ea7d-37f0-51bc-2470-35645d767bbd@comcast.net>, Peter Reilley writes: > >> The chip is run off 12 volts so it must be CMOS. > Or the OCXO is not a 12V model ? > >
SS
Scott Stobbe
Fri, Oct 28, 2016 4:13 PM

The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a
blown regulator in your ocxo, if it is indeed a 12v model.

On Friday, 28 October 2016, Poul-Henning Kamp phk@phk.freebsd.dk wrote:

The chip is run off 12 volts so it must be CMOS.

Or the OCXO is not a 12V model ?

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a blown regulator in your ocxo, if it is indeed a 12v model. On Friday, 28 October 2016, Poul-Henning Kamp <phk@phk.freebsd.dk> wrote: > -------- > In message <10a3ea7d-37f0-51bc-2470-35645d767bbd@comcast.net > <javascript:;>>, Peter Reilley writes: > > >The chip is run off 12 volts so it must be CMOS. > > Or the OCXO is not a 12V model ? > > > -- > Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 > phk@FreeBSD.ORG | TCP/IP since RFC 956 > FreeBSD committer | BSD since 4.3-tahoe > Never attribute to malice what can adequately be explained by incompetence. > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com <javascript:;> > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > and follow the instructions there. >
AG
Adrian Godwin
Fri, Oct 28, 2016 4:27 PM

That's one sweet soldering iron. Is it an American Beauty ?

On Fri, Oct 28, 2016 at 4:31 PM, Peter Reilley preilley_454@comcast.net
wrote:

I did finally get it open.  I used a very large old style soldering
iron and .003 inch steel shim stock.  I would melt the solder on the
straight seams and insert small pieces of the shim.  Solder does not
stick well to steel so the shim kept the soldered seam open.
I used a soldering iron rather than a torch because I can control the
temperature.

I could not use the shim at the corners.  After all the straight seams
were separated I could pull each corner using a screw in the mounting
hole and melt the solder at the corner.  Slowly working my way around,
corner by corner, I got it opened.  I did not damage anything so I
should be able to close it up after I fix it.

Looking around with my scope it seems that the output driver chip is bad
as I expected.  It is a TI 14 pin surface mount DIP.  It says S30 on it
which if it is a 74S30 it is an 8 input positive NAND gate.  The board
layout confirms this as the 10 MHz signal is connected to pin 2 and all
other inputs are tied high.  Pin 8 is connected to the output.

The chip is run off 12 volts so it must be CMOS.  But I cannot find any
chip like that that will run off 12 volts.  Any suggestions for a
replacement?

Also, using an 8 input NAND chip for a driver seems an odd choice.

When I put 12 volts on the unit the S30 chip gets really hot. After I
removed the chip the unit seems to work OK.  The current jumps between
about .1 amp to .9 amps.  It seems like the temperature regulator is
an on/off type controller.

The device on eBay, item 261920574725, looks exactly like what I have.

I have placed a bunch of pictures in my dropbox.
https://www.dropbox.com/sh/52e9d1rva9kpb3w/AABmbIj1aK7Zk2J9SNMmu-JAa?dl=0

Pete.

On 10/18/2016 10:57 AM, J. L. Trantham wrote:

Pete,

I'm not familiar with your OCXO but I found one shown on 'theBay' (item
261920574725) and it appeared to have an option for 'mounting screws', four
of them, on the bottom.  Interestingly, the 'link' to the datasheet for
that unit did not show threads for mounting screws.

If your unit has that option, I would suggest placing four long screws,
mounting the item in a vise, use a small torch (I've used a hand held
propane torch turned down very low to open a number of units from 5061A's)
around the bottom of the case while gripping the top with an appropriate
sized Channel Lock plier and lifting off the top.

If you can repair the OCXO, it should be easy to reassemble the unit with
solder.

TheBay unit looks like it has a screw cover (which likely has a rubber
gasket) for mechanical adjustment of the frequency.  I'd remove that before
applying the torch. :^).

If you get it open, I'd love to see some pictures of the insides.

Good luck and hope this helps.

Joe

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Peter
Reilley
Sent: Tuesday, October 18, 2016 8:11 AM
To: time-nuts@febo.com
Subject: [time-nuts] Opening an Isotemp OCXO

I bought an Isotemp OCXO82-59 with a frequency of 10 MHz for a $3 at the
MIT flea market.
As expected it was dead.  It heats up as expected but looking at the
output with a scope there
is nothing.  However looking at the output with a spectrum analyzer I
can see a faint 10 MHz
signal.  It seems that the oscillator is running but the output
circuitry is dead.  Reasonable
assumption?

Anyway, has anyone had any luck unsoldering the tin case without
destroying it?

Pete.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

That's one sweet soldering iron. Is it an American Beauty ? On Fri, Oct 28, 2016 at 4:31 PM, Peter Reilley <preilley_454@comcast.net> wrote: > I did finally get it open. I used a very large old style soldering > iron and .003 inch steel shim stock. I would melt the solder on the > straight seams and insert small pieces of the shim. Solder does not > stick well to steel so the shim kept the soldered seam open. > I used a soldering iron rather than a torch because I can control the > temperature. > > I could not use the shim at the corners. After all the straight seams > were separated I could pull each corner using a screw in the mounting > hole and melt the solder at the corner. Slowly working my way around, > corner by corner, I got it opened. I did not damage anything so I > should be able to close it up after I fix it. > > Looking around with my scope it seems that the output driver chip is bad > as I expected. It is a TI 14 pin surface mount DIP. It says S30 on it > which if it is a 74S30 it is an 8 input positive NAND gate. The board > layout confirms this as the 10 MHz signal is connected to pin 2 and all > other inputs are tied high. Pin 8 is connected to the output. > > The chip is run off 12 volts so it must be CMOS. But I cannot find any > chip like that that will run off 12 volts. Any suggestions for a > replacement? > > Also, using an 8 input NAND chip for a driver seems an odd choice. > > When I put 12 volts on the unit the S30 chip gets really hot. After I > removed the chip the unit seems to work OK. The current jumps between > about .1 amp to .9 amps. It seems like the temperature regulator is > an on/off type controller. > > The device on eBay, item 261920574725, looks exactly like what I have. > > I have placed a bunch of pictures in my dropbox. > https://www.dropbox.com/sh/52e9d1rva9kpb3w/AABmbIj1aK7Zk2J9SNMmu-JAa?dl=0 > > Pete. > > > > > On 10/18/2016 10:57 AM, J. L. Trantham wrote: > >> Pete, >> >> I'm not familiar with your OCXO but I found one shown on 'theBay' (item >> 261920574725) and it appeared to have an option for 'mounting screws', four >> of them, on the bottom. Interestingly, the 'link' to the datasheet for >> that unit did not show threads for mounting screws. >> >> If your unit has that option, I would suggest placing four long screws, >> mounting the item in a vise, use a small torch (I've used a hand held >> propane torch turned down very low to open a number of units from 5061A's) >> around the bottom of the case while gripping the top with an appropriate >> sized Channel Lock plier and lifting off the top. >> >> If you can repair the OCXO, it should be easy to reassemble the unit with >> solder. >> >> TheBay unit looks like it has a screw cover (which likely has a rubber >> gasket) for mechanical adjustment of the frequency. I'd remove that before >> applying the torch. :^). >> >> If you get it open, I'd love to see some pictures of the insides. >> >> Good luck and hope this helps. >> >> Joe >> >> >> >> -----Original Message----- >> From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Peter >> Reilley >> Sent: Tuesday, October 18, 2016 8:11 AM >> To: time-nuts@febo.com >> Subject: [time-nuts] Opening an Isotemp OCXO >> >> I bought an Isotemp OCXO82-59 with a frequency of 10 MHz for a $3 at the >> MIT flea market. >> As expected it was dead. It heats up as expected but looking at the >> output with a scope there >> is nothing. However looking at the output with a spectrum analyzer I >> can see a faint 10 MHz >> signal. It seems that the oscillator is running but the output >> circuitry is dead. Reasonable >> assumption? >> >> Anyway, has anyone had any luck unsoldering the tin case without >> destroying it? >> >> Pete. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> >> >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/m > ailman/listinfo/time-nuts > and follow the instructions there. >
TM
Tom Miller
Fri, Oct 28, 2016 4:41 PM

When I have an unknown OCXO, I put a scope on the output and connect a bench
power supply to the DC inputs. I bring up the voltage until the RF out hits
a stable level. That is where the internal regulator starts regulating. Then
set it to the nearest normal power supply voltage, +5, +12, +15, +24, +28
volts.

Some OCXOs will have separate inputs for the oscillator and the oven.
Example- an HP OCXO has a +15 regulated input that is switched for the
oscillator and a +24 volt input for the oven that is on always. I usually
try the same voltage as the oscillator first and watch the current as the
oven warms up. If it takes too long to heat up, try a little higher voltage.

Some units have a separate ground return for the oscillator and oven so you
need to watch that.

Good luck and thanks for the report.

Regards,
Tom

----- Original Message -----
From: "Peter Reilley" preilley_454@comcast.net
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Sent: Friday, October 28, 2016 12:09 PM
Subject: Re: [time-nuts] Opening an Isotemp OCXO

The only document that I could find said 12 volt.

Pete.

On 10/28/2016 11:49 AM, Poul-Henning Kamp wrote:


In message 10a3ea7d-37f0-51bc-2470-35645d767bbd@comcast.net, Peter
Reilley writes:

The chip is run off 12 volts so it must be CMOS.

Or the OCXO is not a 12V model ?


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

When I have an unknown OCXO, I put a scope on the output and connect a bench power supply to the DC inputs. I bring up the voltage until the RF out hits a stable level. That is where the internal regulator starts regulating. Then set it to the nearest normal power supply voltage, +5, +12, +15, +24, +28 volts. Some OCXOs will have separate inputs for the oscillator and the oven. Example- an HP OCXO has a +15 regulated input that is switched for the oscillator and a +24 volt input for the oven that is on always. I usually try the same voltage as the oscillator first and watch the current as the oven warms up. If it takes too long to heat up, try a little higher voltage. Some units have a separate ground return for the oscillator and oven so you need to watch that. Good luck and thanks for the report. Regards, Tom ----- Original Message ----- From: "Peter Reilley" <preilley_454@comcast.net> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com> Sent: Friday, October 28, 2016 12:09 PM Subject: Re: [time-nuts] Opening an Isotemp OCXO > The only document that I could find said 12 volt. > > Pete. > > > On 10/28/2016 11:49 AM, Poul-Henning Kamp wrote: >> -------- >> In message <10a3ea7d-37f0-51bc-2470-35645d767bbd@comcast.net>, Peter >> Reilley writes: >> >>> The chip is run off 12 volts so it must be CMOS. >> Or the OCXO is not a 12V model ? >> >> > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
DR
Dan Rae
Fri, Oct 28, 2016 4:42 PM

On 10/28/2016 9:09 AM, Peter Reilley wrote:

The only document that I could find said 12 volt.

Pete.

The 82 series came in a lot of flavors.  I have one 82-49 which is
definitely 12V ( draws 0.12A when warm and maybe 0.3 when cold) and has
a 5V pp square wave output at 10 MHz.  It also has a Vref output and an
EFC input.  What is odd about this one is that it came to me from a TRW
swap meet scrap dealer who bought stuff from Cubic Communications in San
Diego and it had a "Reject" tag from them. Probably because it was
supposed to be a 5 MHz unit as marked, so presumably had the wrong
crystal fitted by Isotemp.  Other versions of the 82 series (-10) were
fitted to Racal Receivers like the RA6790/GM.  Screw adjustment on the
side and 5 MHz output.

Are you sure yours shouldn't have a 5V supply on that output IC? Could
be why it died if there is an internal regulator that failed putting the
full supply voltage on the IC?

Dan

On 10/28/2016 9:09 AM, Peter Reilley wrote: > The only document that I could find said 12 volt. > > Pete. The 82 series came in a lot of flavors. I have one 82-49 which is definitely 12V ( draws 0.12A when warm and maybe 0.3 when cold) and has a 5V pp square wave output at 10 MHz. It also has a Vref output and an EFC input. What is odd about this one is that it came to me from a TRW swap meet scrap dealer who bought stuff from Cubic Communications in San Diego and it had a "Reject" tag from them. Probably because it was supposed to be a 5 MHz unit as marked, so presumably had the wrong crystal fitted by Isotemp. Other versions of the 82 series (-10) were fitted to Racal Receivers like the RA6790/GM. Screw adjustment on the side and 5 MHz output. Are you sure yours shouldn't have a 5V supply on that output IC? Could be why it died if there is an internal regulator that failed putting the full supply voltage on the IC? Dan
G/
Graham / KE9H
Fri, Oct 28, 2016 4:46 PM

Here is the TI document on "Case Marking."  It may not be a 74S30.

http://www.ti.com/lit/an/snoa039c/snoa039c.pdf

--- Graham

==

On Fri, Oct 28, 2016 at 11:27 AM, Adrian Godwin artgodwin@gmail.com wrote:

That's one sweet soldering iron. Is it an American Beauty ?

On Fri, Oct 28, 2016 at 4:31 PM, Peter Reilley preilley_454@comcast.net
wrote:

I did finally get it open.  I used a very large old style soldering
iron and .003 inch steel shim stock.  I would melt the solder on the
straight seams and insert small pieces of the shim.  Solder does not
stick well to steel so the shim kept the soldered seam open.
I used a soldering iron rather than a torch because I can control the
temperature.

I could not use the shim at the corners.  After all the straight seams
were separated I could pull each corner using a screw in the mounting
hole and melt the solder at the corner.  Slowly working my way around,
corner by corner, I got it opened.  I did not damage anything so I
should be able to close it up after I fix it.

Looking around with my scope it seems that the output driver chip is bad
as I expected.  It is a TI 14 pin surface mount DIP.  It says S30 on it
which if it is a 74S30 it is an 8 input positive NAND gate.  The board
layout confirms this as the 10 MHz signal is connected to pin 2 and all
other inputs are tied high.  Pin 8 is connected to the output.

The chip is run off 12 volts so it must be CMOS.  But I cannot find any
chip like that that will run off 12 volts.  Any suggestions for a
replacement?

Also, using an 8 input NAND chip for a driver seems an odd choice.

When I put 12 volts on the unit the S30 chip gets really hot. After I
removed the chip the unit seems to work OK.  The current jumps between
about .1 amp to .9 amps.  It seems like the temperature regulator is
an on/off type controller.

The device on eBay, item 261920574725, looks exactly like what I have.

I have placed a bunch of pictures in my dropbox.
https://www.dropbox.com/sh/52e9d1rva9kpb3w/

AABmbIj1aK7Zk2J9SNMmu-JAa?dl=0

Pete.

On 10/18/2016 10:57 AM, J. L. Trantham wrote:

Pete,

I'm not familiar with your OCXO but I found one shown on 'theBay' (item
261920574725) and it appeared to have an option for 'mounting screws',

four

of them, on the bottom.  Interestingly, the 'link' to the datasheet for
that unit did not show threads for mounting screws.

If your unit has that option, I would suggest placing four long screws,
mounting the item in a vise, use a small torch (I've used a hand held
propane torch turned down very low to open a number of units from

5061A's)

around the bottom of the case while gripping the top with an appropriate
sized Channel Lock plier and lifting off the top.

If you can repair the OCXO, it should be easy to reassemble the unit

with

solder.

TheBay unit looks like it has a screw cover (which likely has a rubber
gasket) for mechanical adjustment of the frequency.  I'd remove that

before

applying the torch. :^).

If you get it open, I'd love to see some pictures of the insides.

Good luck and hope this helps.

Joe

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Peter
Reilley
Sent: Tuesday, October 18, 2016 8:11 AM
To: time-nuts@febo.com
Subject: [time-nuts] Opening an Isotemp OCXO

I bought an Isotemp OCXO82-59 with a frequency of 10 MHz for a $3 at the
MIT flea market.
As expected it was dead.  It heats up as expected but looking at the
output with a scope there
is nothing.  However looking at the output with a spectrum analyzer I
can see a faint 10 MHz
signal.  It seems that the oscillator is running but the output
circuitry is dead.  Reasonable
assumption?

Anyway, has anyone had any luck unsoldering the tin case without
destroying it?

Pete.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

Here is the TI document on "Case Marking." It may not be a 74S30. http://www.ti.com/lit/an/snoa039c/snoa039c.pdf --- Graham == On Fri, Oct 28, 2016 at 11:27 AM, Adrian Godwin <artgodwin@gmail.com> wrote: > That's one sweet soldering iron. Is it an American Beauty ? > > > On Fri, Oct 28, 2016 at 4:31 PM, Peter Reilley <preilley_454@comcast.net> > wrote: > > > I did finally get it open. I used a very large old style soldering > > iron and .003 inch steel shim stock. I would melt the solder on the > > straight seams and insert small pieces of the shim. Solder does not > > stick well to steel so the shim kept the soldered seam open. > > I used a soldering iron rather than a torch because I can control the > > temperature. > > > > I could not use the shim at the corners. After all the straight seams > > were separated I could pull each corner using a screw in the mounting > > hole and melt the solder at the corner. Slowly working my way around, > > corner by corner, I got it opened. I did not damage anything so I > > should be able to close it up after I fix it. > > > > Looking around with my scope it seems that the output driver chip is bad > > as I expected. It is a TI 14 pin surface mount DIP. It says S30 on it > > which if it is a 74S30 it is an 8 input positive NAND gate. The board > > layout confirms this as the 10 MHz signal is connected to pin 2 and all > > other inputs are tied high. Pin 8 is connected to the output. > > > > The chip is run off 12 volts so it must be CMOS. But I cannot find any > > chip like that that will run off 12 volts. Any suggestions for a > > replacement? > > > > Also, using an 8 input NAND chip for a driver seems an odd choice. > > > > When I put 12 volts on the unit the S30 chip gets really hot. After I > > removed the chip the unit seems to work OK. The current jumps between > > about .1 amp to .9 amps. It seems like the temperature regulator is > > an on/off type controller. > > > > The device on eBay, item 261920574725, looks exactly like what I have. > > > > I have placed a bunch of pictures in my dropbox. > > https://www.dropbox.com/sh/52e9d1rva9kpb3w/ > AABmbIj1aK7Zk2J9SNMmu-JAa?dl=0 > > > > Pete. > > > > > > > > > > On 10/18/2016 10:57 AM, J. L. Trantham wrote: > > > >> Pete, > >> > >> I'm not familiar with your OCXO but I found one shown on 'theBay' (item > >> 261920574725) and it appeared to have an option for 'mounting screws', > four > >> of them, on the bottom. Interestingly, the 'link' to the datasheet for > >> that unit did not show threads for mounting screws. > >> > >> If your unit has that option, I would suggest placing four long screws, > >> mounting the item in a vise, use a small torch (I've used a hand held > >> propane torch turned down very low to open a number of units from > 5061A's) > >> around the bottom of the case while gripping the top with an appropriate > >> sized Channel Lock plier and lifting off the top. > >> > >> If you can repair the OCXO, it should be easy to reassemble the unit > with > >> solder. > >> > >> TheBay unit looks like it has a screw cover (which likely has a rubber > >> gasket) for mechanical adjustment of the frequency. I'd remove that > before > >> applying the torch. :^). > >> > >> If you get it open, I'd love to see some pictures of the insides. > >> > >> Good luck and hope this helps. > >> > >> Joe > >> > >> > >> > >> -----Original Message----- > >> From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Peter > >> Reilley > >> Sent: Tuesday, October 18, 2016 8:11 AM > >> To: time-nuts@febo.com > >> Subject: [time-nuts] Opening an Isotemp OCXO > >> > >> I bought an Isotemp OCXO82-59 with a frequency of 10 MHz for a $3 at the > >> MIT flea market. > >> As expected it was dead. It heats up as expected but looking at the > >> output with a scope there > >> is nothing. However looking at the output with a spectrum analyzer I > >> can see a faint 10 MHz > >> signal. It seems that the oscillator is running but the output > >> circuitry is dead. Reasonable > >> assumption? > >> > >> Anyway, has anyone had any luck unsoldering the tin case without > >> destroying it? > >> > >> Pete. > >> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to > >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> and follow the instructions there. > >> > >> > >> > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/m > > ailman/listinfo/time-nuts > > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > and follow the instructions there. >
J
jimlux
Fri, Oct 28, 2016 4:46 PM

On 10/28/16 8:31 AM, Peter Reilley wrote:

Looking around with my scope it seems that the output driver chip is bad
as I expected.  It is a TI 14 pin surface mount DIP.  It says S30 on it
which if it is a 74S30 it is an 8 input positive NAND gate.  The board
layout confirms this as the 10 MHz signal is connected to pin 2 and all
other inputs are tied high.  Pin 8 is connected to the output.

The chip is run off 12 volts so it must be CMOS.  But I cannot find any
chip like that that will run off 12 volts.  Any suggestions for a
replacement?

Also, using an 8 input NAND chip for a driver seems an odd choice.

could it be some sort of 74C series? if it's intended to run off 12V, a
lot of the CMOS parts can run at almost any Vdd (a virtue of 4000 series
CMOS - 15V power, no problem)

Maybe they got a good price on the 8 input NAND parts?

If unused pins are tied high, it's unlikely that they are outputs - most
parts are either a totem pole or open collector output, neither one of
which would like being tied high.

On 10/28/16 8:31 AM, Peter Reilley wrote: > Looking around with my scope it seems that the output driver chip is bad > as I expected. It is a TI 14 pin surface mount DIP. It says S30 on it > which if it is a 74S30 it is an 8 input positive NAND gate. The board > layout confirms this as the 10 MHz signal is connected to pin 2 and all > other inputs are tied high. Pin 8 is connected to the output. > > The chip is run off 12 volts so it must be CMOS. But I cannot find any > chip like that that will run off 12 volts. Any suggestions for a > replacement? > > Also, using an 8 input NAND chip for a driver seems an odd choice. > could it be some sort of 74C series? if it's intended to run off 12V, a lot of the CMOS parts can run at almost any Vdd (a virtue of 4000 series CMOS - 15V power, no problem) Maybe they got a good price on the 8 input NAND parts? If unused pins are tied high, it's unlikely that they are outputs - most parts are either a totem pole or open collector output, neither one of which would like being tied high.
J
jimlux
Fri, Oct 28, 2016 4:48 PM

On 10/28/16 9:13 AM, Scott Stobbe wrote:

The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a
blown regulator in your ocxo, if it is indeed a 12v model.

There you go..the design could use a 74S30 as a driver - it's fast,
fairly good drive, but runs off 5V.  If the regulator is shorted, and
you put 12V on it, it will cook.

On 10/28/16 9:13 AM, Scott Stobbe wrote: > The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a > blown regulator in your ocxo, if it is indeed a 12v model. > There you go..the design could use a 74S30 as a driver - it's fast, fairly good drive, but runs off 5V. If the regulator is shorted, and you put 12V on it, it will cook.
PS
paul swed
Fri, Oct 28, 2016 4:53 PM

I confirmed the pin out matches a 74s30 also. An S30 is TTL. Great pix to
look at.
So 12 V on a 5 V chip is indeed a smoker. Find out why there was 12 V.
OK crazy talk I see a 1K resistor next to the VCC chip. Would anyone be
crazy enough to use a dropping resistor from 12 V to get 5?? Really bad
engineering and I don't actually believe they would. But if true a open
74s30 would indeed show 12 V on pin 14.
Good luck.
Paul
WB8TSL

On Fri, Oct 28, 2016 at 12:48 PM, jimlux jimlux@earthlink.net wrote:

On 10/28/16 9:13 AM, Scott Stobbe wrote:

The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a
blown regulator in your ocxo, if it is indeed a 12v model.

There you go..the design could use a 74S30 as a driver - it's fast,

fairly good drive, but runs off 5V.  If the regulator is shorted, and you
put 12V on it, it will cook.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

I confirmed the pin out matches a 74s30 also. An S30 is TTL. Great pix to look at. So 12 V on a 5 V chip is indeed a smoker. Find out why there was 12 V. OK crazy talk I see a 1K resistor next to the VCC chip. Would anyone be crazy enough to use a dropping resistor from 12 V to get 5?? Really bad engineering and I don't actually believe they would. But if true a open 74s30 would indeed show 12 V on pin 14. Good luck. Paul WB8TSL On Fri, Oct 28, 2016 at 12:48 PM, jimlux <jimlux@earthlink.net> wrote: > On 10/28/16 9:13 AM, Scott Stobbe wrote: > >> The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a >> blown regulator in your ocxo, if it is indeed a 12v model. >> >> There you go..the design could use a 74S30 as a driver - it's fast, > fairly good drive, but runs off 5V. If the regulator is shorted, and you > put 12V on it, it will cook. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/m > ailman/listinfo/time-nuts > and follow the instructions there. >
SS
Scott Stobbe
Fri, Oct 28, 2016 5:03 PM

Could also be a quirk about the 74S30 that gives it better phase noise over
a basic buffer.

On Friday, 28 October 2016, jimlux jimlux@earthlink.net wrote:

On 10/28/16 9:13 AM, Scott Stobbe wrote:

The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a
blown regulator in your ocxo, if it is indeed a 12v model.

There you go..the design could use a 74S30 as a driver - it's fast,

fairly good drive, but runs off 5V.  If the regulator is shorted, and you
put 12V on it, it will cook.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

Could also be a quirk about the 74S30 that gives it better phase noise over a basic buffer. On Friday, 28 October 2016, jimlux <jimlux@earthlink.net> wrote: > On 10/28/16 9:13 AM, Scott Stobbe wrote: > >> The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a >> blown regulator in your ocxo, if it is indeed a 12v model. >> >> There you go..the design could use a 74S30 as a driver - it's fast, > fairly good drive, but runs off 5V. If the regulator is shorted, and you > put 12V on it, it will cook. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/m > ailman/listinfo/time-nuts > and follow the instructions there. >
AG
Adrian Godwin
Fri, Oct 28, 2016 5:03 PM

The 1k resistor doesn't seem to feed the 'S30'. It looks as though pin 14
(Vcc) goes via that thick track to the +12 input.

On Fri, Oct 28, 2016 at 5:46 PM, jimlux jimlux@earthlink.net wrote:

On 10/28/16 8:31 AM, Peter Reilley wrote:

Looking around with my scope it seems that the output driver chip is bad

as I expected.  It is a TI 14 pin surface mount DIP.  It says S30 on it
which if it is a 74S30 it is an 8 input positive NAND gate.  The board
layout confirms this as the 10 MHz signal is connected to pin 2 and all
other inputs are tied high.  Pin 8 is connected to the output.

The chip is run off 12 volts so it must be CMOS.  But I cannot find any
chip like that that will run off 12 volts.  Any suggestions for a
replacement?

Also, using an 8 input NAND chip for a driver seems an odd choice.

could it be some sort of 74C series? if it's intended to run off 12V, a
lot of the CMOS parts can run at almost any Vdd (a virtue of 4000 series
CMOS - 15V power, no problem)

Maybe they got a good price on the 8 input NAND parts?

If unused pins are tied high, it's unlikely that they are outputs - most
parts are either a totem pole or open collector output, neither one of
which would like being tied high.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

The 1k resistor doesn't seem to feed the 'S30'. It looks as though pin 14 (Vcc) goes via that thick track to the +12 input. On Fri, Oct 28, 2016 at 5:46 PM, jimlux <jimlux@earthlink.net> wrote: > On 10/28/16 8:31 AM, Peter Reilley wrote: > > Looking around with my scope it seems that the output driver chip is bad >> as I expected. It is a TI 14 pin surface mount DIP. It says S30 on it >> which if it is a 74S30 it is an 8 input positive NAND gate. The board >> layout confirms this as the 10 MHz signal is connected to pin 2 and all >> other inputs are tied high. Pin 8 is connected to the output. >> >> The chip is run off 12 volts so it must be CMOS. But I cannot find any >> chip like that that will run off 12 volts. Any suggestions for a >> replacement? >> >> Also, using an 8 input NAND chip for a driver seems an odd choice. >> >> > could it be some sort of 74C series? if it's intended to run off 12V, a > lot of the CMOS parts can run at almost any Vdd (a virtue of 4000 series > CMOS - 15V power, no problem) > > Maybe they got a good price on the 8 input NAND parts? > > If unused pins are tied high, it's unlikely that they are outputs - most > parts are either a totem pole or open collector output, neither one of > which would like being tied high. > > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/m > ailman/listinfo/time-nuts > and follow the instructions there. >
SS
Scott Stobbe
Fri, Oct 28, 2016 5:31 PM

I wounder if originally the designer was hoping to use all 8 wire or'd
inputs to lower the input referred noise during midscale transition. Then
backed out later for some reason.

On Fri, Oct 28, 2016 at 1:03 PM, Scott Stobbe scott.j.stobbe@gmail.com
wrote:

Could also be a quirk about the 74S30 that gives it better phase noise
over a basic buffer.

On Friday, 28 October 2016, jimlux jimlux@earthlink.net wrote:

On 10/28/16 9:13 AM, Scott Stobbe wrote:

The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a
blown regulator in your ocxo, if it is indeed a 12v model.

There you go..the design could use a 74S30 as a driver - it's fast,

fairly good drive, but runs off 5V.  If the regulator is shorted, and you
put 12V on it, it will cook.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

I wounder if originally the designer was hoping to use all 8 wire or'd inputs to lower the input referred noise during midscale transition. Then backed out later for some reason. On Fri, Oct 28, 2016 at 1:03 PM, Scott Stobbe <scott.j.stobbe@gmail.com> wrote: > Could also be a quirk about the 74S30 that gives it better phase noise > over a basic buffer. > > > On Friday, 28 October 2016, jimlux <jimlux@earthlink.net> wrote: > >> On 10/28/16 9:13 AM, Scott Stobbe wrote: >> >>> The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a >>> blown regulator in your ocxo, if it is indeed a 12v model. >>> >>> There you go..the design could use a 74S30 as a driver - it's fast, >> fairly good drive, but runs off 5V. If the regulator is shorted, and you >> put 12V on it, it will cook. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/m >> ailman/listinfo/time-nuts >> and follow the instructions there. >> >
PR
Peter Reilley
Fri, Oct 28, 2016 7:07 PM

There is no regulator chip in the unit.  I am thinking that this must
be a 5 volt unit.

Pete.

On 10/28/2016 12:42 PM, Dan Rae wrote:

On 10/28/2016 9:09 AM, Peter Reilley wrote:

The only document that I could find said 12 volt.

Pete.

The 82 series came in a lot of flavors.  I have one 82-49 which is
definitely 12V ( draws 0.12A when warm and maybe 0.3 when cold) and
has a 5V pp square wave output at 10 MHz.  It also has a Vref output
and an EFC input.  What is odd about this one is that it came to me
from a TRW swap meet scrap dealer who bought stuff from Cubic
Communications in San Diego and it had a "Reject" tag from them.
Probably because it was supposed to be a 5 MHz unit as marked, so
presumably had the wrong crystal fitted by Isotemp. Other versions of
the 82 series (-10) were fitted to Racal Receivers like the
RA6790/GM.  Screw adjustment on the side and 5 MHz output.

Are you sure yours shouldn't have a 5V supply on that output IC? Could
be why it died if there is an internal regulator that failed putting
the full supply voltage on the IC?

Dan

There is no regulator chip in the unit. I am thinking that this must be a 5 volt unit. Pete. On 10/28/2016 12:42 PM, Dan Rae wrote: > On 10/28/2016 9:09 AM, Peter Reilley wrote: >> The only document that I could find said 12 volt. >> >> Pete. > The 82 series came in a lot of flavors. I have one 82-49 which is > definitely 12V ( draws 0.12A when warm and maybe 0.3 when cold) and > has a 5V pp square wave output at 10 MHz. It also has a Vref output > and an EFC input. What is odd about this one is that it came to me > from a TRW swap meet scrap dealer who bought stuff from Cubic > Communications in San Diego and it had a "Reject" tag from them. > Probably because it was supposed to be a 5 MHz unit as marked, so > presumably had the wrong crystal fitted by Isotemp. Other versions of > the 82 series (-10) were fitted to Racal Receivers like the > RA6790/GM. Screw adjustment on the side and 5 MHz output. > > Are you sure yours shouldn't have a 5V supply on that output IC? Could > be why it died if there is an internal regulator that failed putting > the full supply voltage on the IC? > > Dan > > >
PR
Peter Reilley
Fri, Oct 28, 2016 7:10 PM

The reason that there was 12 volts on the unit was because I put it
there.  I should
have tried 5 volts first but the only datasheet that I could find said
12 volts.
All the eBay units that look the same say 12 volts.

Pete.

On 10/28/2016 12:53 PM, paul swed wrote:

I confirmed the pin out matches a 74s30 also. An S30 is TTL. Great pix to
look at.
So 12 V on a 5 V chip is indeed a smoker. Find out why there was 12 V.
OK crazy talk I see a 1K resistor next to the VCC chip. Would anyone be
crazy enough to use a dropping resistor from 12 V to get 5?? Really bad
engineering and I don't actually believe they would. But if true a open
74s30 would indeed show 12 V on pin 14.
Good luck.
Paul
WB8TSL

On Fri, Oct 28, 2016 at 12:48 PM, jimlux jimlux@earthlink.net wrote:

On 10/28/16 9:13 AM, Scott Stobbe wrote:

The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a
blown regulator in your ocxo, if it is indeed a 12v model.

There you go..the design could use a 74S30 as a driver - it's fast,

fairly good drive, but runs off 5V.  If the regulator is shorted, and you
put 12V on it, it will cook.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

The reason that there was 12 volts on the unit was because I put it there. I should have tried 5 volts first but the only datasheet that I could find said 12 volts. All the eBay units that look the same say 12 volts. Pete. On 10/28/2016 12:53 PM, paul swed wrote: > I confirmed the pin out matches a 74s30 also. An S30 is TTL. Great pix to > look at. > So 12 V on a 5 V chip is indeed a smoker. Find out why there was 12 V. > OK crazy talk I see a 1K resistor next to the VCC chip. Would anyone be > crazy enough to use a dropping resistor from 12 V to get 5?? Really bad > engineering and I don't actually believe they would. But if true a open > 74s30 would indeed show 12 V on pin 14. > Good luck. > Paul > WB8TSL > > On Fri, Oct 28, 2016 at 12:48 PM, jimlux <jimlux@earthlink.net> wrote: > >> On 10/28/16 9:13 AM, Scott Stobbe wrote: >> >>> The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a >>> blown regulator in your ocxo, if it is indeed a 12v model. >>> >>> There you go..the design could use a 74S30 as a driver - it's fast, >> fairly good drive, but runs off 5V. If the regulator is shorted, and you >> put 12V on it, it will cook. >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/m >> ailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
PR
Peter Reilley
Fri, Oct 28, 2016 7:14 PM

I don't know, I looked for a name and could not find one.
Sometimes there is no substitute for a big block of copper when soldering.
I pick them up at flea markets, no one wants them.  I have a few.

Pete.

On 10/28/2016 12:27 PM, Adrian Godwin wrote:

That's one sweet soldering iron. Is it an American Beauty ?

I don't know, I looked for a name and could not find one. Sometimes there is no substitute for a big block of copper when soldering. I pick them up at flea markets, no one wants them. I have a few. Pete. On 10/28/2016 12:27 PM, Adrian Godwin wrote: > That's one sweet soldering iron. Is it an American Beauty ? > >
AG
Adrian Godwin
Fri, Oct 28, 2016 7:18 PM

The data sheet for the device only mentions 12V. Other data sheets (eg the
131) mention 5V and 12V, but don't indicate how to specify one or the
other.

Gerry Sweeny's article at
http://gerrysweeney.com/diy-hpagilent-53131a-010-high-stability-timebase-option/
mentions a part number suffix that distinguishes two choices but this
doesn't appear in the main datasheet. Other writers here have in the past
mentioned that the suffix has no predefined meaning and is used only to
distinguish design variants internally.

It appears that Isotemp do offer a choice of voltage options but don't like
to document it publicly. Tom Miller's strategy seems the safest if using a
surplus device.

On Fri, Oct 28, 2016 at 8:10 PM, Peter Reilley preilley_454@comcast.net
wrote:

The reason that there was 12 volts on the unit was because I put it
there.  I should
have tried 5 volts first but the only datasheet that I could find said 12
volts.
All the eBay units that look the same say 12 volts.

Pete.

On 10/28/2016 12:53 PM, paul swed wrote:

I confirmed the pin out matches a 74s30 also. An S30 is TTL. Great pix to
look at.
So 12 V on a 5 V chip is indeed a smoker. Find out why there was 12 V.
OK crazy talk I see a 1K resistor next to the VCC chip. Would anyone be
crazy enough to use a dropping resistor from 12 V to get 5?? Really bad
engineering and I don't actually believe they would. But if true a open
74s30 would indeed show 12 V on pin 14.
Good luck.
Paul
WB8TSL

On Fri, Oct 28, 2016 at 12:48 PM, jimlux jimlux@earthlink.net wrote:

On 10/28/16 9:13 AM, Scott Stobbe wrote:

The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a

blown regulator in your ocxo, if it is indeed a 12v model.

There you go..the design could use a 74S30 as a driver - it's fast,

fairly good drive, but runs off 5V.  If the regulator is shorted, and you
put 12V on it, it will cook.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

The data sheet for the device only mentions 12V. Other data sheets (eg the 131) mention 5V and 12V, but don't indicate how to specify one or the other. Gerry Sweeny's article at http://gerrysweeney.com/diy-hpagilent-53131a-010-high-stability-timebase-option/ mentions a part number suffix that distinguishes two choices but this doesn't appear in the main datasheet. Other writers here have in the past mentioned that the suffix has no predefined meaning and is used only to distinguish design variants internally. It appears that Isotemp do offer a choice of voltage options but don't like to document it publicly. Tom Miller's strategy seems the safest if using a surplus device. On Fri, Oct 28, 2016 at 8:10 PM, Peter Reilley <preilley_454@comcast.net> wrote: > The reason that there was 12 volts on the unit was because I put it > there. I should > have tried 5 volts first but the only datasheet that I could find said 12 > volts. > All the eBay units that look the same say 12 volts. > > Pete. > > > On 10/28/2016 12:53 PM, paul swed wrote: > >> I confirmed the pin out matches a 74s30 also. An S30 is TTL. Great pix to >> look at. >> So 12 V on a 5 V chip is indeed a smoker. Find out why there was 12 V. >> OK crazy talk I see a 1K resistor next to the VCC chip. Would anyone be >> crazy enough to use a dropping resistor from 12 V to get 5?? Really bad >> engineering and I don't actually believe they would. But if true a open >> 74s30 would indeed show 12 V on pin 14. >> Good luck. >> Paul >> WB8TSL >> >> On Fri, Oct 28, 2016 at 12:48 PM, jimlux <jimlux@earthlink.net> wrote: >> >> On 10/28/16 9:13 AM, Scott Stobbe wrote: >>> >>> The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a >>>> blown regulator in your ocxo, if it is indeed a 12v model. >>>> >>>> There you go..the design could use a 74S30 as a driver - it's fast, >>>> >>> fairly good drive, but runs off 5V. If the regulator is shorted, and you >>> put 12V on it, it will cook. >>> >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/m >>> ailman/listinfo/time-nuts >>> and follow the instructions there. >>> >>> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/m >> ailman/listinfo/time-nuts >> and follow the instructions there. >> >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/m > ailman/listinfo/time-nuts > and follow the instructions there. >
TM
Tom Miller
Fri, Oct 28, 2016 7:20 PM

It looks like that is the only device that could be damaged by 12 volts. Can
you find a replacement and try running at 5 volts?

----- Original Message -----
From: "Peter Reilley" preilley_454@comcast.net
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Sent: Friday, October 28, 2016 3:10 PM
Subject: Re: [time-nuts] Opening an Isotemp OCXO

The reason that there was 12 volts on the unit was because I put it there.
I should
have tried 5 volts first but the only datasheet that I could find said 12
volts.
All the eBay units that look the same say 12 volts.

Pete.

On 10/28/2016 12:53 PM, paul swed wrote:

I confirmed the pin out matches a 74s30 also. An S30 is TTL. Great pix to
look at.
So 12 V on a 5 V chip is indeed a smoker. Find out why there was 12 V.
OK crazy talk I see a 1K resistor next to the VCC chip. Would anyone be
crazy enough to use a dropping resistor from 12 V to get 5?? Really bad
engineering and I don't actually believe they would. But if true a open
74s30 would indeed show 12 V on pin 14.
Good luck.
Paul
WB8TSL

On Fri, Oct 28, 2016 at 12:48 PM, jimlux jimlux@earthlink.net wrote:

On 10/28/16 9:13 AM, Scott Stobbe wrote:

The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a
blown regulator in your ocxo, if it is indeed a 12v model.

There you go..the design could use a 74S30 as a driver - it's fast,

fairly good drive, but runs off 5V.  If the regulator is shorted, and
you
put 12V on it, it will cook.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

It looks like that is the only device that could be damaged by 12 volts. Can you find a replacement and try running at 5 volts? ----- Original Message ----- From: "Peter Reilley" <preilley_454@comcast.net> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com> Sent: Friday, October 28, 2016 3:10 PM Subject: Re: [time-nuts] Opening an Isotemp OCXO > The reason that there was 12 volts on the unit was because I put it there. > I should > have tried 5 volts first but the only datasheet that I could find said 12 > volts. > All the eBay units that look the same say 12 volts. > > Pete. > > On 10/28/2016 12:53 PM, paul swed wrote: >> I confirmed the pin out matches a 74s30 also. An S30 is TTL. Great pix to >> look at. >> So 12 V on a 5 V chip is indeed a smoker. Find out why there was 12 V. >> OK crazy talk I see a 1K resistor next to the VCC chip. Would anyone be >> crazy enough to use a dropping resistor from 12 V to get 5?? Really bad >> engineering and I don't actually believe they would. But if true a open >> 74s30 would indeed show 12 V on pin 14. >> Good luck. >> Paul >> WB8TSL >> >> On Fri, Oct 28, 2016 at 12:48 PM, jimlux <jimlux@earthlink.net> wrote: >> >>> On 10/28/16 9:13 AM, Scott Stobbe wrote: >>> >>>> The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a >>>> blown regulator in your ocxo, if it is indeed a 12v model. >>>> >>>> There you go..the design could use a 74S30 as a driver - it's fast, >>> fairly good drive, but runs off 5V. If the regulator is shorted, and >>> you >>> put 12V on it, it will cook. >>> >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/m >>> ailman/listinfo/time-nuts >>> and follow the instructions there. >>> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BC
Bob Camp
Fri, Oct 28, 2016 7:46 PM

Hi

Roughly 99.9% of all OCXO’s made go to large OEM customers. The percentage may
actually be a bit higher than that. There are relatively few markets that “catalog” OCXO’s
sell into.

Inevitably the first thing that an OEM wants is some form of customization. A specific
supply voltage, a certain output format, a wider (or narrower) EFC range … there
are lots of possibilities. For every OCXO that goes into production for these guys,
five or ten other designs are done (all equally custom) that never see the light of day
past the samples.

The spec’s that these parts are built to are negotiated between the supplier and the OEM.
In some cases they are the property of the OEM and the spec is their control drawing on
the part. In other cases the drawing is done by the supplier for that OEM and is property
of the supplier.

The OEM often has competitors. They would love to get access to the OEM’s control
drawings to see how the systems are designed. The supplier has competitors. They
would love to get access to the suppliers drawings so they can make cheap knock off
parts to those drawings. In both cases, the drawings (in general) have very real value.

The net result of this is that both suppliers and OEM’s put fairly fancy rules in place about
passing out drawings. More or less anything up to and including being boiled in oil is
(if legal in the jurisdiction) fair recourse under most of these rules. Needless to say
people learn pretty quick that you get fired for this sort of thing.

The net result is that the drawings for most OCXO’s simply do not exist in the public
domain. They do (or did) exist in some form somewhere. Getting at public copies of
them is highly unlikely. Going by “similar looking” drawings is not a real good idea ….

Bob

On Oct 28, 2016, at 3:14 PM, Peter Reilley preilley_454@comcast.net wrote:

I don't know, I looked for a name and could not find one.
Sometimes there is no substitute for a big block of copper when soldering.
I pick them up at flea markets, no one wants them.  I have a few.

Pete.

On 10/28/2016 12:27 PM, Adrian Godwin wrote:

That's one sweet soldering iron. Is it an American Beauty ?


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Roughly 99.9% of all OCXO’s made go to large OEM customers. The percentage may actually be a bit higher than that. There are relatively few markets that “catalog” OCXO’s sell into. Inevitably the first thing that an OEM wants is some form of customization. A specific supply voltage, a certain output format, a wider (or narrower) EFC range … there are lots of possibilities. For every OCXO that goes into production for these guys, five or ten other designs are done (all equally custom) that never see the light of day past the samples. The spec’s that these parts are built to are negotiated between the supplier and the OEM. In some cases they are the property of the OEM and the spec is their control drawing on the part. In other cases the drawing is done by the supplier for that OEM and is property of the supplier. The OEM often has competitors. They would *love* to get access to the OEM’s control drawings to see how the systems are designed. The supplier has competitors. They would love to get access to the suppliers drawings so they can make cheap knock off parts to those drawings. In both cases, the drawings (in general) have very real value. The net result of this is that both suppliers and OEM’s put fairly fancy rules in place about passing out drawings. More or less anything up to and including being boiled in oil is (if legal in the jurisdiction) fair recourse under most of these rules. Needless to say people learn pretty quick that you get fired for this sort of thing. The net result is that the drawings for most OCXO’s simply do not exist in the public domain. They do (or did) exist in some form somewhere. Getting at public copies of them is highly unlikely. Going by “similar looking” drawings is not a real good idea …. Bob > On Oct 28, 2016, at 3:14 PM, Peter Reilley <preilley_454@comcast.net> wrote: > > I don't know, I looked for a name and could not find one. > Sometimes there is no substitute for a big block of copper when soldering. > I pick them up at flea markets, no one wants them. I have a few. > > Pete. > > On 10/28/2016 12:27 PM, Adrian Godwin wrote: >> That's one sweet soldering iron. Is it an American Beauty ? >> >> > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BS
Bob Stewart
Fri, Oct 28, 2016 8:09 PM

Hi Bob,
Can't the OCXOs be characterized pretty closely by someone with the right tools and staff?  I don't have a big sample to speak from, but the Trimbles I use only have a couple of ceramic coated pieces, and those can be exposed down to the die by hand and then characterized, can't they?  Granted, that's beyond my capabilities, but given the right tools and employees...

Bob

  From: Bob Camp <kb8tq@n1k.org>

To: peter@reilley.com; Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Friday, October 28, 2016 2:46 PM
Subject: Re: [time-nuts] Opening an Isotemp OCXO

Hi

Roughly 99.9% of all OCXO’s made go to large OEM customers. The percentage may
actually be a bit higher than that. There are relatively few markets that “catalog” OCXO’s
sell into.

Inevitably the first thing that an OEM wants is some form of customization. A specific
supply voltage, a certain output format, a wider (or narrower) EFC range … there
are lots of possibilities. For every OCXO that goes into production for these guys,
five or ten other designs are done (all equally custom) that never see the light of day
past the samples.

The spec’s that these parts are built to are negotiated between the supplier and the OEM.
In some cases they are the property of the OEM and the spec is their control drawing on
the part. In other cases the drawing is done by the supplier for that OEM and is property
of the supplier.

The OEM often has competitors. They would love to get access to the OEM’s control
drawings to see how the systems are designed. The supplier has competitors. They
would love to get access to the suppliers drawings so they can make cheap knock off
parts to those drawings. In both cases, the drawings (in general) have very real value.

The net result of this is that both suppliers and OEM’s put fairly fancy rules in place about
passing out drawings. More or less anything up to and including being boiled in oil is
(if legal in the jurisdiction) fair recourse under most of these rules. Needless to say
people learn pretty quick that you get fired for this sort of thing.

The net result is that the drawings for most OCXO’s simply do not exist in the public
domain. They do (or did) exist in some form somewhere. Getting at public copies of
them is highly unlikely. Going by “similar looking” drawings is not a real good idea ….

Bob

Hi Bob, Can't the OCXOs be characterized pretty closely by someone with the right tools and staff?  I don't have a big sample to speak from, but the Trimbles I use only have a couple of ceramic coated pieces, and those can be exposed down to the die by hand and then characterized, can't they?  Granted, that's beyond my capabilities, but given the right tools and employees... Bob From: Bob Camp <kb8tq@n1k.org> To: peter@reilley.com; Discussion of precise time and frequency measurement <time-nuts@febo.com> Sent: Friday, October 28, 2016 2:46 PM Subject: Re: [time-nuts] Opening an Isotemp OCXO Hi Roughly 99.9% of all OCXO’s made go to large OEM customers. The percentage may actually be a bit higher than that. There are relatively few markets that “catalog” OCXO’s sell into. Inevitably the first thing that an OEM wants is some form of customization. A specific supply voltage, a certain output format, a wider (or narrower) EFC range … there are lots of possibilities. For every OCXO that goes into production for these guys, five or ten other designs are done (all equally custom) that never see the light of day past the samples. The spec’s that these parts are built to are negotiated between the supplier and the OEM. In some cases they are the property of the OEM and the spec is their control drawing on the part. In other cases the drawing is done by the supplier for that OEM and is property of the supplier. The OEM often has competitors. They would *love* to get access to the OEM’s control drawings to see how the systems are designed. The supplier has competitors. They would love to get access to the suppliers drawings so they can make cheap knock off parts to those drawings. In both cases, the drawings (in general) have very real value. The net result of this is that both suppliers and OEM’s put fairly fancy rules in place about passing out drawings. More or less anything up to and including being boiled in oil is (if legal in the jurisdiction) fair recourse under most of these rules. Needless to say people learn pretty quick that you get fired for this sort of thing. The net result is that the drawings for most OCXO’s simply do not exist in the public domain. They do (or did) exist in some form somewhere. Getting at public copies of them is highly unlikely. Going by “similar looking” drawings is not a real good idea …. Bob
BS
Bob Stewart
Fri, Oct 28, 2016 8:14 PM

From experience, I can tell you that you can't always depend on the seller when it comes to how to feed your new OCXO.  For instance, the seller of the Trimble 65256 insisted that it took 12V.  So, that's what I used, and all the magic smoke came tumbling out.

Bob
 
From: Peter Reilley preilley_454@comcast.net
To: Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Friday, October 28, 2016 2:10 PM
Subject: Re: [time-nuts] Opening an Isotemp OCXO

The reason that there was 12 volts on the unit was because I put it
there.  I should
have tried 5 volts first but the only datasheet that I could find said
12 volts.
All the eBay units that look the same say 12 volts.

Pete.

>From experience, I can tell you that you can't always depend on the seller when it comes to how to feed your new OCXO.  For instance, the seller of the Trimble 65256 insisted that it took 12V.  So, that's what I used, and all the magic smoke came tumbling out. Bob   From: Peter Reilley <preilley_454@comcast.net> To: Discussion of precise time and frequency measurement <time-nuts@febo.com> Sent: Friday, October 28, 2016 2:10 PM Subject: Re: [time-nuts] Opening an Isotemp OCXO The reason that there was 12 volts on the unit was because I put it there.  I should have tried 5 volts first but the only datasheet that I could find said 12 volts. All the eBay units that look the same say 12 volts. Pete.
D
David
Fri, Oct 28, 2016 8:36 PM

I do not think the designer was considering noise at all because tying
the inputs together would not do anything useful.  Emitter resistance
is inversely proportional to emitter current (26/mA) but putting them
in parallel lowers the current through each emitter so the total
emitter resistance stays the same.

Supply current is separate for each TTL gate so by using a single
8-input part, total power it is about half that of a dual 4-input part
and a quarter of a quad 2-input part depending on the exact operating
conditions.  Unused outputs should be high for lowest power.

74S30 Single 8-Input 5mA 10mA
74S20 Dual 4-Input 8mA 18mA
74S00 Quad 2-Input 16mA 36mA

One clever design I ran across used the 7420 dual 4-input NAND pinout
but wired the inputs which are pin compatible so the 7400 quad 2-input
pinout would also work.  This allows using a 74x00, 74x20, or 74S120
dual 4-input NAND buffer but it draws even more power.

74S120 Dual 4-Input 18mA 44mA

On Fri, 28 Oct 2016 13:31:38 -0400, you wrote:

I wounder if originally the designer was hoping to use all 8 wire or'd
inputs to lower the input referred noise during midscale transition. Then
backed out later for some reason.

On Fri, Oct 28, 2016 at 1:03 PM, Scott Stobbe scott.j.stobbe@gmail.com
wrote:

Could also be a quirk about the 74S30 that gives it better phase noise
over a basic buffer.

On Friday, 28 October 2016, jimlux jimlux@earthlink.net wrote:

On 10/28/16 9:13 AM, Scott Stobbe wrote:

The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a
blown regulator in your ocxo, if it is indeed a 12v model.

There you go..the design could use a 74S30 as a driver - it's fast,

fairly good drive, but runs off 5V.  If the regulator is shorted, and you
put 12V on it, it will cook.

I do not think the designer was considering noise at all because tying the inputs together would not do anything useful. Emitter resistance is inversely proportional to emitter current (26/mA) but putting them in parallel lowers the current through each emitter so the total emitter resistance stays the same. Supply current is separate for each TTL gate so by using a single 8-input part, total power it is about half that of a dual 4-input part and a quarter of a quad 2-input part depending on the exact operating conditions. Unused *outputs* should be high for lowest power. 74S30 Single 8-Input 5mA 10mA 74S20 Dual 4-Input 8mA 18mA 74S00 Quad 2-Input 16mA 36mA One clever design I ran across used the 7420 dual 4-input NAND pinout but wired the inputs which are pin compatible so the 7400 quad 2-input pinout would also work. This allows using a 74x00, 74x20, or 74S120 dual 4-input NAND buffer but it draws even more power. 74S120 Dual 4-Input 18mA 44mA On Fri, 28 Oct 2016 13:31:38 -0400, you wrote: >I wounder if originally the designer was hoping to use all 8 wire or'd >inputs to lower the input referred noise during midscale transition. Then >backed out later for some reason. > >On Fri, Oct 28, 2016 at 1:03 PM, Scott Stobbe <scott.j.stobbe@gmail.com> >wrote: > >> Could also be a quirk about the 74S30 that gives it better phase noise >> over a basic buffer. >> >> >> On Friday, 28 October 2016, jimlux <jimlux@earthlink.net> wrote: >> >>> On 10/28/16 9:13 AM, Scott Stobbe wrote: >>> >>>> The OCXO82-59 datasheet lists 12V supply, 5V clock out, could also be a >>>> blown regulator in your ocxo, if it is indeed a 12v model. >>>> >>>> There you go..the design could use a 74S30 as a driver - it's fast, >>> fairly good drive, but runs off 5V. If the regulator is shorted, and you >>> put 12V on it, it will cook.
BC
Bob Camp
Fri, Oct 28, 2016 9:09 PM

Hi

Well, the first issue would be the one that started this thread…. what’s the supply and pinout.
Yes, you can tear a couple open and work that out. Next step would be to verify that the other
couple hundred in the batch all are the same pinout and supply.

Assuming that you have all the same parts, next up would be making the test fixtures for the
units. It does not need to be super complex, but you need a way to solidly connect to the part.
If the leads have been cut off, this is a bit more complex than if you have full length leads.

Most processes age the units first. That would involve putting all hundred pieces on power and
looking at each of them with a counter every couple of minutes for a month or two. Based on that
data you could get a pretty good idea of what the aging will run. You also will weed out some
percentage of the units that didn’t survive whatever process got them to you.

After stabilization, They would go into temperature test. Likely something like a dozen or two
per run. Since you don’t know the top end, I’d do a search for that first. I’d then do a search for
the low end. Based on data from a few runs (several dozen parts) you should have some idea
of the upper and lower temp limits.

A formal temp run over that range would be next. I’d probably do 5 degree steps. That way if
you are off on the endpoint guess you might be able to see the correct end points. You would
test the whole batch of units and then look at the data. My guess is that you would re-guess the
end points and re-test the batch at that point.

Assuming you know the correct load and EFC from the original tear down, they would move on
to some sort of bench check. If not, you would need to work that out.

On the bench check, I’d run each one over the (assumed) EFC range at something like 0.1V steps.
Here if your range is off, the data probably is still ok. That should give you a proper EFC setting
for each one. Since they have been aged and TC’d at this point, the EFC center point should
be pretty good.

After that, things like phase noise and ADEV would be on the list. Same thing, run them all
and see what they do. Make some decisions and toss out the outliers.

So, yes it can be done. Because of the tear down process early on and the data redundancy
needed, you have to get a pretty good sized group of units. The risk is that you get a group
of parts with a common (to that batch) defect. They were headed to the garbage dump and
went to eBay instead ….

Lots of Fun

Bob

On Oct 28, 2016, at 4:09 PM, Bob Stewart bob@evoria.net wrote:

Hi Bob,
Can't the OCXOs be characterized pretty closely by someone with the right tools and staff?  I don't have a big sample to speak from, but the Trimbles I use only have a couple of ceramic coated pieces, and those can be exposed down to the die by hand and then characterized, can't they?  Granted, that's beyond my capabilities, but given the right tools and employees...

Bob

  From: Bob Camp <kb8tq@n1k.org>

To: peter@reilley.com; Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Friday, October 28, 2016 2:46 PM
Subject: Re: [time-nuts] Opening an Isotemp OCXO

Hi

Roughly 99.9% of all OCXO’s made go to large OEM customers. The percentage may
actually be a bit higher than that. There are relatively few markets that “catalog” OCXO’s
sell into.

Inevitably the first thing that an OEM wants is some form of customization. A specific
supply voltage, a certain output format, a wider (or narrower) EFC range … there
are lots of possibilities. For every OCXO that goes into production for these guys,
five or ten other designs are done (all equally custom) that never see the light of day
past the samples.

The spec’s that these parts are built to are negotiated between the supplier and the OEM.
In some cases they are the property of the OEM and the spec is their control drawing on
the part. In other cases the drawing is done by the supplier for that OEM and is property
of the supplier.

The OEM often has competitors. They would love to get access to the OEM’s control
drawings to see how the systems are designed. The supplier has competitors. They
would love to get access to the suppliers drawings so they can make cheap knock off
parts to those drawings. In both cases, the drawings (in general) have very real value.

The net result of this is that both suppliers and OEM’s put fairly fancy rules in place about
passing out drawings. More or less anything up to and including being boiled in oil is
(if legal in the jurisdiction) fair recourse under most of these rules. Needless to say
people learn pretty quick that you get fired for this sort of thing.

The net result is that the drawings for most OCXO’s simply do not exist in the public
domain. They do (or did) exist in some form somewhere. Getting at public copies of
them is highly unlikely. Going by “similar looking” drawings is not a real good idea ….

Bob


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Well, the first issue would be the one that started this thread…. what’s the supply and pinout. Yes, you can tear a couple open and work that out. Next step would be to verify that the other couple hundred in the batch all are the same pinout and supply. Assuming that you have all the same parts, next up would be making the test fixtures for the units. It does not need to be super complex, but you need a way to solidly connect to the part. If the leads have been cut off, this is a bit more complex than if you have full length leads. Most processes age the units first. That would involve putting all hundred pieces on power and looking at each of them with a counter every couple of minutes for a month or two. Based on that data you could get a pretty good idea of what the aging will run. You also will weed out some percentage of the units that didn’t survive whatever process got them to you. After stabilization, They would go into temperature test. Likely something like a dozen or two per run. Since you don’t know the top end, I’d do a search for that first. I’d then do a search for the low end. Based on data from a few runs (several dozen parts) you should have some idea of the upper and lower temp limits. A formal temp run over that range would be next. I’d probably do 5 degree steps. That way if you are off on the endpoint guess you might be able to see the correct end points. You would test the whole batch of units and then look at the data. My guess is that you would re-guess the end points and re-test the batch at that point. Assuming you know the correct load and EFC from the original tear down, they would move on to some sort of bench check. If not, you would need to work that out. On the bench check, I’d run each one over the (assumed) EFC range at something like 0.1V steps. Here if your range is off, the data probably is still ok. That should give you a proper EFC setting for each one. Since they have been aged and TC’d at this point, the EFC center point should be pretty good. After that, things like phase noise and ADEV would be on the list. Same thing, run them all and see what they do. Make some decisions and toss out the outliers. So, yes it can be done. Because of the tear down process early on and the data redundancy needed, you have to get a pretty good sized group of units. The risk is that you get a group of parts with a common (to that batch) defect. They *were* headed to the garbage dump and went to eBay instead …. Lots of Fun Bob > On Oct 28, 2016, at 4:09 PM, Bob Stewart <bob@evoria.net> wrote: > > Hi Bob, > Can't the OCXOs be characterized pretty closely by someone with the right tools and staff? I don't have a big sample to speak from, but the Trimbles I use only have a couple of ceramic coated pieces, and those can be exposed down to the die by hand and then characterized, can't they? Granted, that's beyond my capabilities, but given the right tools and employees... > > Bob > > From: Bob Camp <kb8tq@n1k.org> > To: peter@reilley.com; Discussion of precise time and frequency measurement <time-nuts@febo.com> > Sent: Friday, October 28, 2016 2:46 PM > Subject: Re: [time-nuts] Opening an Isotemp OCXO > > Hi > > Roughly 99.9% of all OCXO’s made go to large OEM customers. The percentage may > actually be a bit higher than that. There are relatively few markets that “catalog” OCXO’s > sell into. > > Inevitably the first thing that an OEM wants is some form of customization. A specific > supply voltage, a certain output format, a wider (or narrower) EFC range … there > are lots of possibilities. For every OCXO that goes into production for these guys, > five or ten other designs are done (all equally custom) that never see the light of day > past the samples. > > The spec’s that these parts are built to are negotiated between the supplier and the OEM. > In some cases they are the property of the OEM and the spec is their control drawing on > the part. In other cases the drawing is done by the supplier for that OEM and is property > of the supplier. > > The OEM often has competitors. They would *love* to get access to the OEM’s control > drawings to see how the systems are designed. The supplier has competitors. They > would love to get access to the suppliers drawings so they can make cheap knock off > parts to those drawings. In both cases, the drawings (in general) have very real value. > > The net result of this is that both suppliers and OEM’s put fairly fancy rules in place about > passing out drawings. More or less anything up to and including being boiled in oil is > (if legal in the jurisdiction) fair recourse under most of these rules. Needless to say > people learn pretty quick that you get fired for this sort of thing. > > The net result is that the drawings for most OCXO’s simply do not exist in the public > domain. They do (or did) exist in some form somewhere. Getting at public copies of > them is highly unlikely. Going by “similar looking” drawings is not a real good idea …. > > Bob > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
J
jimlux
Fri, Oct 28, 2016 9:59 PM

On 10/28/16 1:09 PM, Bob Stewart wrote:

Hi Bob,
Can't the OCXOs be characterized pretty closely by someone with the right tools and staff?  I don't have a big sample to speak from, but the Trimbles I use only have a couple of ceramic coated pieces, and those can be exposed down to the die by hand and then characterized, can't they?  Granted, that's beyond my capabilities, but given the right tools and employees...

The real thing is "which requirement do I really, really care about"
beyond the straightforward frequency stability/phase noise specs.

You might have a specific "I care about 1 Hz to 100 Hz, but don't much
are about >100 Hz or <1Hz" and you're willing to let the mfr do pretty
much anything else for the rest.

I recently had a requirement where I don't much care about absolute
frequency accuracy, but I do care about phase noise and short term (<3
seconds) stability.  We got some quotes for OCXOs with the oven
disabled, figuring that we don't need to burn the power for the oven,
since that's more about frequency accuracy, and our environment (in
space) doesn't vary more than a couple degrees over hours.  That's
something you'll never see in a catalog.

We buy oscillators at JPL for things like landing radars - the operating
life is minutes, but it had better work for those minutes - we care not
what the aging or frequency accuracy is in this situation.

I know people that have bought parts for telemetry transmitters on
devices with life measured in seconds - there, the critical requirement
was "must start from -80C and run for 10 seconds"

A coworker was telling me about requirements for oscillators in fire
control radars attached to gatling guns - Now there's a vibe sensitivity
requirement.

So, as Bob pointed out, most oscillators go into applications that have
idiosyncratic requirements and a cost or delivery schedule requirement
pushing against the overall performance requirement.  Not everyone wants
a 1kg ultra stable oscillator from APL with a 3 year delivery time.

On 10/28/16 1:09 PM, Bob Stewart wrote: > Hi Bob, > Can't the OCXOs be characterized pretty closely by someone with the right tools and staff? I don't have a big sample to speak from, but the Trimbles I use only have a couple of ceramic coated pieces, and those can be exposed down to the die by hand and then characterized, can't they? Granted, that's beyond my capabilities, but given the right tools and employees... > The real thing is "which requirement do I really, really care about" beyond the straightforward frequency stability/phase noise specs. You might have a specific "I care about 1 Hz to 100 Hz, but don't much are about >100 Hz or <1Hz" and you're willing to let the mfr do pretty much anything else for the rest. I recently had a requirement where I don't much care about absolute frequency accuracy, but I do care about phase noise and short term (<3 seconds) stability. We got some quotes for OCXOs with the oven disabled, figuring that we don't need to burn the power for the oven, since that's more about frequency accuracy, and our environment (in space) doesn't vary more than a couple degrees over hours. That's something you'll never see in a catalog. We buy oscillators at JPL for things like landing radars - the operating life is minutes, but it had better work for those minutes - we care not what the aging or frequency accuracy is in this situation. I know people that have bought parts for telemetry transmitters on devices with life measured in seconds - there, the critical requirement was "must start from -80C and run for 10 seconds" A coworker was telling me about requirements for oscillators in fire control radars attached to gatling guns - Now there's a vibe sensitivity requirement. So, as Bob pointed out, most oscillators go into applications that have idiosyncratic requirements and a cost or delivery schedule requirement pushing against the overall performance requirement. Not everyone wants a 1kg ultra stable oscillator from APL with a 3 year delivery time.
PR
Peter Reilley
Sat, Oct 29, 2016 1:51 PM

More information;

I added a picture to the dropbox from my Flir IR camera.  The picture
shows the copper block
that the crystal is attached to running at about 200 F.  In the IR shot
the copper block is to the
right.    In most of the regular pictures it is toward the bottom of the
picture.    This is with the
unit (minus the S30 chip) running on 5 volts for more than 10 hours.
Is that too hot?

While running at 5 volts the current is constant at about 4. amps, no
cycling.  At 12 volts
it cycled between .9 to .1 amps.  I would not expect cycling for the
temperature control
of an OCXO.  I would expect a linear temperature control circuit.

I looked at the tantalum capacitor on the bottom of the board.  The
marking is 39-10.
Does that mean 39 uF and 10 volts?  If so then it must be a 5 volt
unit.  The capacitor
did not explode at 12 volts.

Dropbox link:
https://www.dropbox.com/sh/52e9d1rva9kpb3w/AABmbIj1aK7Zk2J9SNMmu-JAa?dl=0

Pete.

On 10/18/2016 9:11 AM, Peter Reilley wrote:

I bought an Isotemp OCXO82-59 with a frequency of 10 MHz for a $3 at
the MIT flea market.
As expected it was dead.  It heats up as expected but looking at the
output with a scope there
is nothing.  However looking at the output with a spectrum analyzer I
can see a faint 10 MHz
signal.  It seems that the oscillator is running but the output
circuitry is dead.  Reasonable
assumption?

Anyway, has anyone had any luck unsoldering the tin case without
destroying it?

Pete.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

More information; I added a picture to the dropbox from my Flir IR camera. The picture shows the copper block that the crystal is attached to running at about 200 F. In the IR shot the copper block is to the right. In most of the regular pictures it is toward the bottom of the picture. This is with the unit (minus the S30 chip) running on 5 volts for more than 10 hours. Is that too hot? While running at 5 volts the current is constant at about 4. amps, no cycling. At 12 volts it cycled between .9 to .1 amps. I would not expect cycling for the temperature control of an OCXO. I would expect a linear temperature control circuit. I looked at the tantalum capacitor on the bottom of the board. The marking is 39-10. Does that mean 39 uF and 10 volts? If so then it must be a 5 volt unit. The capacitor did not explode at 12 volts. Dropbox link: https://www.dropbox.com/sh/52e9d1rva9kpb3w/AABmbIj1aK7Zk2J9SNMmu-JAa?dl=0 Pete. On 10/18/2016 9:11 AM, Peter Reilley wrote: > I bought an Isotemp OCXO82-59 with a frequency of 10 MHz for a $3 at > the MIT flea market. > As expected it was dead. It heats up as expected but looking at the > output with a scope there > is nothing. However looking at the output with a spectrum analyzer I > can see a faint 10 MHz > signal. It seems that the oscillator is running but the output > circuitry is dead. Reasonable > assumption? > > Anyway, has anyone had any luck unsoldering the tin case without > destroying it? > > Pete. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >