Greg Burnett wrote:
Joe,
I think 1 o'clock (instead of 12 o'clock) might still be OK and/or typical?
Maybe some other 5370A users can tell us where their Trigger Level pots are
pointing when centered for triggering on a weak signal. (The pots point to
12 o'clock on my 5370B, but I can't remember exactly where they might point
on a typical "A".)
My 5370A triggers on the frequency standard output on either the START
or STOP inputs set to 10x attenuation when the Level setting knobs are
at the ~1 oclock position.
The average of the threshold settings when the trigger level is just
above and just below the point at which the trigger light flashes is
20mV or less.
Not a particularly precise measure as it depends on the waveform symmetry.
Both channels trigger (10x input attenuation switched in) on the
frequency standard output when the the threshold is preset.
It is unfortunate that their is no readout of the EXT arm input trigger
level although there is a trigger LED.
I do know that some of the 5370A A3/A4 Input Amp & Schmitt Trigger I.C.s
developed "I.C. rot" many years after manufacture (due to a contamination
problem during manufacture). For units so afflicted, the symptom was that
"Preset" trigger level was no longer correct, but the trigger level could
still be successfully set manually. For these cases, as the years progressed
(and the "rot" worsened), you had to turn the manual trigger level pot
further and further to one direction to find trigger. As the years of
deterioration continued, there came a point where both trigger level
setability and sensitivity were no longer acceptable.
If you have only a small trigger level issue in "Preset" mode, my thinking
is that you can still correct for it by doing the A3/A4 input adjustments.
Best,
Greg
Bruce
oh yeah-cool :-)
Don
Joseph Gray
So now all your HP equipment looks like a 1950's parking lot, huh?
On Sat, Oct 10, 2009 at 4:16 PM, Don Latham djl@montana.com wrote:
Hi All: I put extended fins on my 1980's Hp equipment by drilling and
tapping the existing sink. I used the white heatsink grease from Radio
Shack to help.
Don Latham
Joseph Gray
The heatsink on my unit is quite hot, also. I'm thinking about putting
a fan on it after I get everything working.
On Sat, Oct 10, 2009 at 9:24 AM, Greg Burnett gbusg@comcast.net
wrote:
Roy,
The HP 5370A rear mounted heat-sink typically heats up to around 61
degrees
C. This is very hot to the touch, so I suspect your unit's temperature
is
normal (unless it's running way hotter than 61 degrees C).
The buzzing transformer might just be due to loose, vibrating
laminations.
You might try tightening the transformer's bolts to see if that
reduces
the
buzz?
Best,
Greg
P.S. I agree that the rear mounted heat sinks of most other HP
equipment
run
significantly cooler. The models that run so hot to the touch are the
5370A/B, 5359A, and many of the legacy HP pulse generators (from
Boblingen
Division).
----- Original Message -----
From: "Roy Phillips" phill.r1@btinternet.com
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Sent: Saturday, October 10, 2009 4:08 AM
Subject: Re: [time-nuts] 5370A
Hi Joe
I have a similar problem so I would be pleased to hear from you with
any
useful information. The basic problem with mine is that the power
supply
is
suspect - it works, but there is excessive heat, much greater than any
other
HP instrument that I have, from the rear mounted heat-sink, this is
too
hot
to hold your hand on ! There is also a very obvious buzz from the
power
transformer, even when it is on "stand-by" - so I guess at worst it
could be
shorted turns in the power transformer, or probably better, a faulty
bridge
rectifier,or one or more of the large reservoir caps - this an early
model
(1984), so perhaps it is to be suspected. I also think the performance
is
not to specification, but I will check-out the perceived problem with
PSU
before I investigate this matter. I am just about to start the
investigation of the prime problem.
Roy
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
--
Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
--
Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com
I plan on doing the A3/A4 adjustments. I just don't know where I'm
going to get a pulse generator or the fancy scope with the 1GHz
sampling plugin. Not the type of equipment on your average test bench.
You won't need any exotic gear for the DAC voltage adjustments or the A3/A4
input assembly adjustments, which is what comes first in the manual. As I
recall the scope is only used for symmetry adjustments in that section of
the manual, so the risetime is not all that important as long as it will
give you a stable trigger.
You can skip the pulse-response adjustment unless you have reason to suspect
a specific problem. Watch for the pots that this step calls for, and don't
center them at the beginning of the alignment process when the manual tells
you to. As I recall there are various typos in the manual, so watch out for
that too. :) In general, the 5370 alignment instructions are crap by HP's
usual standards, so be careful. Read and understand everything before
doing anything.
The A18 DAC adjustments should actually be done first IMHO. Don't worry if
the test limit voltages are not achievable -- they aren't, on any 'B' model
I've ever aligned -- as long as they are symmetrical. For example, +/- 1.9
volts is OK, instead of the +/- 2.2 volts called for in some versions of the
manual.
If you don't have an 8082A pulse generator and a fast scope, you will want
to stop there, and not mess with A19/A20. Likewise you probably do not need
to align the A21 multiplier (which is the only step that requires a spectrum
analyzer) if you don't have reason to suspect a problem with it. I believe
some people have found that the standard-deviation floor can be improved by
realigning A21 but I haven't done that myself.
The instructions that call for a 1 GHz sampling scope don't actually need a
sampling scope; they can be followed with a Tek 2467-class scope as well.
But again, the front-end work should be doable with a good 100 MHz+ model,
and the rest should not be attempted without an 8082A pulse generator.
The nasty thing about aligning a 5370 is the fact that the steps generally
begin with setting all the pots and trimmers to a fixed position, usually
midpoint. Ordinarily it's best not to disturb the alignment of a stage
that's already in tolerance, but they don't give you that option with the
5370, and you may get pretty far before you discover that your test
equipment such as the pulse generator is not up to the task. When they call
for an 8082A in the A19/A20 section, they mean an 8082A, and you don't
want to get halfway into that procedure before you realize that.
-- john, KE5FX
John,
Thanks for the good advice. I definitely need to pull the front panel
to replace the flakey BNC. After that, I'll do what I can with the
A3/A4 adjustments. I just took another look at what I was doing last
night and the two input channels do need adjusting. The B channel is
worse.
I have a few other things that need to get done this weekend, so the
5370A will be on the bench for a little while. I'll get back to you
after I get the inputs adjusted.
Joe Gray
KA5ZEC
On Sat, Oct 10, 2009 at 6:07 PM, John Miles jmiles@pop.net wrote:
I plan on doing the A3/A4 adjustments. I just don't know where I'm
going to get a pulse generator or the fancy scope with the 1GHz
sampling plugin. Not the type of equipment on your average test bench.
You won't need any exotic gear for the DAC voltage adjustments or the A3/A4
input assembly adjustments, which is what comes first in the manual. As I
recall the scope is only used for symmetry adjustments in that section of
the manual, so the risetime is not all that important as long as it will
give you a stable trigger.
You can skip the pulse-response adjustment unless you have reason to suspect
a specific problem. Watch for the pots that this step calls for, and don't
center them at the beginning of the alignment process when the manual tells
you to. As I recall there are various typos in the manual, so watch out for
that too. :) In general, the 5370 alignment instructions are crap by HP's
usual standards, so be careful. Read and understand everything before
doing anything.
The A18 DAC adjustments should actually be done first IMHO. Don't worry if
the test limit voltages are not achievable -- they aren't, on any 'B' model
I've ever aligned -- as long as they are symmetrical. For example, +/- 1.9
volts is OK, instead of the +/- 2.2 volts called for in some versions of the
manual.
If you don't have an 8082A pulse generator and a fast scope, you will want
to stop there, and not mess with A19/A20. Likewise you probably do not need
to align the A21 multiplier (which is the only step that requires a spectrum
analyzer) if you don't have reason to suspect a problem with it. I believe
some people have found that the standard-deviation floor can be improved by
realigning A21 but I haven't done that myself.
The instructions that call for a 1 GHz sampling scope don't actually need a
sampling scope; they can be followed with a Tek 2467-class scope as well.
But again, the front-end work should be doable with a good 100 MHz+ model,
and the rest should not be attempted without an 8082A pulse generator.
The nasty thing about aligning a 5370 is the fact that the steps generally
begin with setting all the pots and trimmers to a fixed position, usually
midpoint. Ordinarily it's best not to disturb the alignment of a stage
that's already in tolerance, but they don't give you that option with the
5370, and you may get pretty far before you discover that your test
equipment such as the pulse generator is not up to the task. When they call
for an 8082A in the A19/A20 section, they mean an 8082A, and you don't
want to get halfway into that procedure before you realize that.
-- john, KE5FX