Hi all, recently I bought a TAPR TICC as a learning project, and also
bought two TADD-2 Minis to serve as sine-to-square converters for the
two inputs. Since the TADD-2 Minis will be dedicated for this purpose, I
simplified their assembly by just installing a jumper from pin 2 to 3 of
the IC socket position, installed the BNC input and output connectors,
and +12vdc power wiring. (I.e. I omitted the PIC chip, its socket and
the other headers.)

As a test, I sent the 10 MHz out from my Trimble Thunderbolt through one
TADD-2 Mini and to a Siglent digital scope. A screen shot from the scope
is attached.

I have two questions at this point:

  1. Does this look like a decent waveform for TICC purposes? I verified
    in advance with the scope that the Thunderbolt is putting out a clean
    sine wave.

  2. For this test, I just used 50-ohm coax between the Thunderbolt and
    the TADD-2 Mini, and between the latter and the scope. I see the TADD-2
    Mini instructions suggest avoiding a 50-ohm load, instead terminating
    into a TTL or high-impedance load. Is there a recommendation for the
    best cabling to use to connect the TADD-2 Minis to the TICC?

Thanks for any advice,

Frank

Hi all, recently I bought a TAPR TICC as a learning project, and also bought two TADD-2 Minis to serve as sine-to-square converters for the two inputs. Since the TADD-2 Minis will be dedicated for this purpose, I simplified their assembly by just installing a jumper from pin 2 to 3 of the IC socket position, installed the BNC input and output connectors, and +12vdc power wiring. (I.e. I omitted the PIC chip, its socket and the other headers.) As a test, I sent the 10 MHz out from my Trimble Thunderbolt through one TADD-2 Mini and to a Siglent digital scope. A screen shot from the scope is attached. I have two questions at this point: 1) Does this look like a decent waveform for TICC purposes? I verified in advance with the scope that the Thunderbolt is putting out a clean sine wave. 2) For this test, I just used 50-ohm coax between the Thunderbolt and the TADD-2 Mini, and between the latter and the scope. I see the TADD-2 Mini instructions suggest avoiding a 50-ohm load, instead terminating into a TTL or high-impedance load. Is there a recommendation for the best cabling to use to connect the TADD-2 Minis to the TICC? Thanks for any advice, Frank
CS
Charles Steinmetz
Mon, Jun 1, 2020 3:14 AM

Frank wrote:

As a test, I sent the 10 MHz out from my Trimble Thunderbolt through one
TADD-2 Mini and to a Siglent digital scope. A screen shot from the scope
is attached.

  1. Does this look like a decent waveform for TICC purposes? I verified
    in advance with the scope that the Thunderbolt is putting out a clean
    sine wave.

No, that is not a promising waveform for driving any sort of
edge-triggered counting circuitry.  There is way too much ringing.  Note
that there is only about a 1.25v margin from the bottom of the ringing
after the rising edge and the top of the ringing after the falling edge.
Any edge-triggered circuitry would have to reliably trigger within
this rather narrow window under all conditions, automatically.  Not
impossible, but very far from optimum.

  1. For this test, I just used 50-ohm coax between the Thunderbolt and
    the TADD-2 Mini, and between the latter and the scope. I see the TADD-2
    Mini instructions suggest avoiding a 50-ohm load, instead terminating
    into a TTL or high-impedance load. Is there a recommendation for the
    best cabling to use to connect the TADD-2 Minis to the TICC?

No idea why the T2-Mini recommends against using a termination that
matches the coax.  The TAPR blurb on the T2-Mini says, "The T2-Mini has
a single low-impedance output that delivers greater than 3.5 volts into
a 50 ohm load."  That is what you need to do -- use a 50 ohm termination
at the load end of the coax between the T2-Mini and the TICC to tame the
ringing.

Once you hook the T2-Mini up to the TICC, you can do that easily by
installing the jumper at the TICC input (JP3), to put 51 ohm resistor
R17 across the input connector.

Doing as you are now -- feeding the T2-Mini output through coax to the
input of your scope -- you will either need to switch the scope input
impedance from 1 Mohm to 50 ohms (if the scope has that feature --
sadly, many digiscopes do not), or use a 50 ohm feedthrough terminator
on the scope input.

Best regards,

Charles

Frank wrote: > As a test, I sent the 10 MHz out from my Trimble Thunderbolt through one > TADD-2 Mini and to a Siglent digital scope. A screen shot from the scope > is attached. > > 1) Does this look like a decent waveform for TICC purposes? I verified > in advance with the scope that the Thunderbolt is putting out a clean > sine wave. No, that is not a promising waveform for driving any sort of edge-triggered counting circuitry. There is way too much ringing. Note that there is only about a 1.25v margin from the bottom of the ringing after the rising edge and the top of the ringing after the falling edge. Any edge-triggered circuitry would have to reliably trigger within this rather narrow window under all conditions, automatically. Not impossible, but very far from optimum. > 2) For this test, I just used 50-ohm coax between the Thunderbolt and > the TADD-2 Mini, and between the latter and the scope. I see the TADD-2 > Mini instructions suggest avoiding a 50-ohm load, instead terminating > into a TTL or high-impedance load. Is there a recommendation for the > best cabling to use to connect the TADD-2 Minis to the TICC? No idea why the T2-Mini recommends against using a termination that matches the coax. The TAPR blurb on the T2-Mini says, "The T2-Mini has a single low-impedance output that delivers greater than 3.5 volts into a 50 ohm load." That is what you need to do -- use a 50 ohm termination at the load end of the coax between the T2-Mini and the TICC to tame the ringing. Once you hook the T2-Mini up to the TICC, you can do that easily by installing the jumper at the TICC input (JP3), to put 51 ohm resistor R17 across the input connector. Doing as you are now -- feeding the T2-Mini output through coax to the input of your scope -- you will either need to switch the scope input impedance from 1 Mohm to 50 ohms (if the scope has that feature -- sadly, many digiscopes do not), or use a 50 ohm feedthrough terminator on the scope input. Best regards, Charles
FO
Frank O'Donnell
Mon, Jun 1, 2020 6:29 AM

On 5/31/20 8:14 PM, Charles Steinmetz wrote:

Doing as you are now -- feeding the T2-Mini output through coax to the
input of your scope -- you will either need to switch the scope input
impedance from 1 Mohm to 50 ohms (if the scope has that feature --
sadly, many digiscopes do not), or use a 50 ohm feedthrough terminator
on the scope input.

Thanks for the comments. Attached is a screen shot with the
Thunderbolt's 10 MHz output again going through the T2-Mini to the
scope, but with a 50-ohm feed-through terminator placed on the scope's
input.

Frank

On 5/31/20 8:14 PM, Charles Steinmetz wrote: > Doing as you are now -- feeding the T2-Mini output through coax to the > input of your scope -- you will either need to switch the scope input > impedance from 1 Mohm to 50 ohms (if the scope has that feature -- > sadly, many digiscopes do not), or use a 50 ohm feedthrough terminator > on the scope input. Thanks for the comments. Attached is a screen shot with the Thunderbolt's 10 MHz output again going through the T2-Mini to the scope, but with a 50-ohm feed-through terminator placed on the scope's input. Frank
BK
Bob kb8tq
Mon, Jun 1, 2020 4:11 PM

Hi

Scope basics:

It is not unusual for a scope to have an input channel with a capacitance
of 10 to 20 pf. That does not sound like much but ….

A 10 MHz square wave has significant energy at 30, 50, 70, 90 MHz (and on up).
Unless a coax is terminated at all these frequencies they will be reflected and you
will get a ring.

So 20 pf at 10 MHz is 795j ohms. At 90 MHz 88j ohms. A feed through termination
probably does a reasonable job at 10 MHz …. it’s pretty awful  by the time you get
to 90 MHz. It is somewhere in the “ummm …. errr …” range at 30, 50, and 70 MHz.

At 10 MHz, 50 + 795j ohms will give you an 18 db return loss. ( the reflected signal
will be 18 db below the incoming signal). A voltage at 1/10 the input is 20 db down.
Not all of the reflected signal will be “ring”, it has to bounce off the source first. How
much bounces depends a lot on how good the source match is.

Bottom line - for a good 50 ohm termination, you need to do things another way.
One of the reasons there are compensated scope probes is to get around this
issue. There are a variety of them out there with a range of capabilities. Yes, the
fancy fet probes cost more than something simple …….

A 10 db pad ahead of the feed through is another way to deal with this. It should at
lest get you into the 20 db return loss range.

Even scopes that have 50 ohm inputs may or may not have good 50 ohm inputs.
Is that clearly spelled out in the specs? Often it isn’t …. sorry about that.

Bob

On Jun 1, 2020, at 2:29 AM, Frank O'Donnell time@inkbox.net wrote:

On 5/31/20 8:14 PM, Charles Steinmetz wrote:

Doing as you are now -- feeding the T2-Mini output through coax to the input of your scope -- you will either need to switch the scope input impedance from 1 Mohm to 50 ohms (if the scope has that feature -- sadly, many digiscopes do not), or use a 50 ohm feedthrough terminator on the scope input.

Thanks for the comments. Attached is a screen shot with the Thunderbolt's 10 MHz output again going through the T2-Mini to the scope, but with a 50-ohm feed-through terminator placed on the scope's input.

Frank

<tadd2-mini-50ohm-term.jpg>_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi Scope basics: It is not unusual for a scope to have an input channel with a capacitance of 10 to 20 pf. That does not sound like much but …. A 10 MHz square wave has significant energy at 30, 50, 70, 90 MHz (and on up). Unless a coax is terminated at all these frequencies they will be reflected and you will get a ring. So 20 pf at 10 MHz is 795j ohms. At 90 MHz 88j ohms. A feed through termination probably does a reasonable job at 10 MHz …. it’s pretty awful by the time you get to 90 MHz. It is somewhere in the “ummm …. errr …” range at 30, 50, and 70 MHz. At 10 MHz, 50 + 795j ohms will give you an 18 db return loss. ( the reflected signal will be 18 db below the incoming signal). A voltage at 1/10 the input is 20 db down. Not all of the reflected signal will be “ring”, it has to bounce off the source first. How much bounces depends a lot on how good the *source* match is. Bottom line - for a *good* 50 ohm termination, you need to do things another way. One of the reasons there are compensated scope probes is to get around this issue. There are a variety of them out there with a range of capabilities. Yes, the fancy fet probes cost more than something simple ……. A 10 db pad ahead of the feed through is another way to deal with this. It should at lest get you into the 20 db return loss range. Even scopes that *have* 50 ohm inputs may or may not have *good* 50 ohm inputs. Is that clearly spelled out in the specs? Often it isn’t …. sorry about that. Bob > On Jun 1, 2020, at 2:29 AM, Frank O'Donnell <time@inkbox.net> wrote: > > On 5/31/20 8:14 PM, Charles Steinmetz wrote: >> Doing as you are now -- feeding the T2-Mini output through coax to the input of your scope -- you will either need to switch the scope input impedance from 1 Mohm to 50 ohms (if the scope has that feature -- sadly, many digiscopes do not), or use a 50 ohm feedthrough terminator on the scope input. > > Thanks for the comments. Attached is a screen shot with the Thunderbolt's 10 MHz output again going through the T2-Mini to the scope, but with a 50-ohm feed-through terminator placed on the scope's input. > > Frank > > <tadd2-mini-50ohm-term.jpg>_______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
FO
Frank O'Donnell
Tue, Jun 2, 2020 5:18 PM

On 6/1/20 9:11 AM, Bob kb8tq wrote:

Scope basics:

Thanks for the comments. Very helpful as background beyond this immediate project.

A 10 db pad ahead of the feed through is another way to deal with this. It should at
lest get you into the 20 db return loss range.

I have a nice JFW variable attenuator, and inserted it between the TADD-2 Mini and the scope to capture the Thunderbolt's 10 MHz signal converted to square wave with 0, 3, 6, 10 and 20 dB attenuation. Screen shots are attached (very small, but there are five of them).

One of the reasons there are compensated scope probes is to get around this
issue. There are a variety of them out there with a range of capabilities. Yes, the
fancy fet probes cost more than something simple …….

Currently the only probes I have is the stock pair that came with the Siglent digital scope (an SDS1202?X-e). I also recently bought a vintage Tektronix T922 analog scope, but it didn't include probes. Thus I might be in the market for a good set of probes. If there are suggestions for any that aren't more expensive than, say, a T922 or a used Thunderbolt, I'd be very interested.

Thanks again,

Frank

On 6/1/20 9:11 AM, Bob kb8tq wrote: > Scope basics: Thanks for the comments. Very helpful as background beyond this immediate project. > A 10 db pad ahead of the feed through is another way to deal with this. It should at > lest get you into the 20 db return loss range. I have a nice JFW variable attenuator, and inserted it between the TADD-2 Mini and the scope to capture the Thunderbolt's 10 MHz signal converted to square wave with 0, 3, 6, 10 and 20 dB attenuation. Screen shots are attached (very small, but there are five of them). > One of the reasons there are compensated scope probes is to get around this > issue. There are a variety of them out there with a range of capabilities. Yes, the > fancy fet probes cost more than something simple ……. Currently the only probes I have is the stock pair that came with the Siglent digital scope (an SDS1202?X-e). I also recently bought a vintage Tektronix T922 analog scope, but it didn't include probes. Thus I might be in the market for a good set of probes. If there are suggestions for any that aren't more expensive than, say, a T922 or a used Thunderbolt, I'd be very interested. Thanks again, Frank
BK
Bob kb8tq
Tue, Jun 2, 2020 6:01 PM

Hi

There’s a lot more to setting this stuff up past the basics. Not all
attenuators have good return loss …. (and on and on). This quickly
gets pretty far of topic for Time Nuts. Simply keep in mind that the
measurement gear and measurement technique very much get
into the results of almost anything you look at ….

Bob

On Jun 2, 2020, at 1:18 PM, Frank O'Donnell time@inkbox.net wrote:

On 6/1/20 9:11 AM, Bob kb8tq wrote:

Scope basics:

Thanks for the comments. Very helpful as background beyond this immediate project.

A 10 db pad ahead of the feed through is another way to deal with this. It should at
lest get you into the 20 db return loss range.

I have a nice JFW variable attenuator, and inserted it between the TADD-2 Mini and the scope to capture the Thunderbolt's 10 MHz signal converted to square wave with 0, 3, 6, 10 and 20 dB attenuation. Screen shots are attached (very small, but there are five of them).

One of the reasons there are compensated scope probes is to get around this
issue. There are a variety of them out there with a range of capabilities. Yes, the
fancy fet probes cost more than something simple …….

Currently the only probes I have is the stock pair that came with the Siglent digital scope (an SDS1202?X-e). I also recently bought a vintage Tektronix T922 analog scope, but it didn't include probes. Thus I might be in the market for a good set of probes. If there are suggestions for any that aren't more expensive than, say, a T922 or a used Thunderbolt, I'd be very interested.

Thanks again,

Frank

<tadd2mini-50ohm-0db.png><tadd2mini-50ohm-3db.png><tadd2mini-50ohm-6db.png><tadd2mini-50ohm-10db.png><tadd2mini-50ohm-20db.png>_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi There’s a lot more to setting this stuff up past the basics. Not all attenuators have good return loss …. (and on and on). This quickly gets pretty far of topic for Time Nuts. Simply keep in mind that the measurement gear and measurement technique very much get into the results of almost anything you look at …. Bob > On Jun 2, 2020, at 1:18 PM, Frank O'Donnell <time@inkbox.net> wrote: > > On 6/1/20 9:11 AM, Bob kb8tq wrote: > >> Scope basics: > > > Thanks for the comments. Very helpful as background beyond this immediate project. > > >> A 10 db pad ahead of the feed through is another way to deal with this. It should at >> lest get you into the 20 db return loss range. > > > I have a nice JFW variable attenuator, and inserted it between the TADD-2 Mini and the scope to capture the Thunderbolt's 10 MHz signal converted to square wave with 0, 3, 6, 10 and 20 dB attenuation. Screen shots are attached (very small, but there are five of them). > > > >> One of the reasons there are compensated scope probes is to get around this >> issue. There are a variety of them out there with a range of capabilities. Yes, the >> fancy fet probes cost more than something simple ……. > > > Currently the only probes I have is the stock pair that came with the Siglent digital scope (an SDS1202?X-e). I also recently bought a vintage Tektronix T922 analog scope, but it didn't include probes. Thus I might be in the market for a good set of probes. If there are suggestions for any that aren't more expensive than, say, a T922 or a used Thunderbolt, I'd be very interested. > > > Thanks again, > > Frank > > <tadd2mini-50ohm-0db.png><tadd2mini-50ohm-3db.png><tadd2mini-50ohm-6db.png><tadd2mini-50ohm-10db.png><tadd2mini-50ohm-20db.png>_______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
TH
Tom Holmes
Tue, Jun 2, 2020 9:06 PM

Frank...

What is interesting is that all of your traces appear to be identical except for the vertical scale factor, which suggests that the ringing you are seeing, which is much improved from your earlier plots, is still a little bit of a problem. I am going to guess that , as Bob pointed out, is probably due to a reflection problem in the input side of the attenuator. Could be the attenuator input impedance is not a good match to the source, or could be a cable impedance issue.

I hope that suggests some things you could try experimentally and educationally to sort it out.

Tom Holmes, N8ZM

-----Original Message-----
From: time-nuts time-nuts-bounces@lists.febo.com On Behalf Of Frank O'Donnell
Sent: Tuesday, June 02, 2020 1:19 PM
To: time-nuts@lists.febo.com
Subject: Re: [time-nuts] TAPR TICC and TADD-2 Minis

On 6/1/20 9:11 AM, Bob kb8tq wrote:

Scope basics:

Thanks for the comments. Very helpful as background beyond this immediate project.

A 10 db pad ahead of the feed through is another way to deal with this. It should at
lest get you into the 20 db return loss range.

I have a nice JFW variable attenuator, and inserted it between the TADD-2 Mini and the scope to capture the Thunderbolt's 10 MHz signal converted to square wave with 0, 3, 6, 10 and 20 dB attenuation. Screen shots are attached (very small, but there are five of them).

One of the reasons there are compensated scope probes is to get around this
issue. There are a variety of them out there with a range of capabilities. Yes, the
fancy fet probes cost more than something simple …….

Currently the only probes I have is the stock pair that came with the Siglent digital scope (an SDS1202?X-e). I also recently bought a vintage Tektronix T922 analog scope, but it didn't include probes. Thus I might be in the market for a good set of probes. If there are suggestions for any that aren't more expensive than, say, a T922 or a used Thunderbolt, I'd be very interested.

Thanks again,

Frank

Frank... What is interesting is that all of your traces appear to be identical except for the vertical scale factor, which suggests that the ringing you are seeing, which is much improved from your earlier plots, is still a little bit of a problem. I am going to guess that , as Bob pointed out, is probably due to a reflection problem in the input side of the attenuator. Could be the attenuator input impedance is not a good match to the source, or could be a cable impedance issue. I hope that suggests some things you could try experimentally and educationally to sort it out. Tom Holmes, N8ZM -----Original Message----- From: time-nuts <time-nuts-bounces@lists.febo.com> On Behalf Of Frank O'Donnell Sent: Tuesday, June 02, 2020 1:19 PM To: time-nuts@lists.febo.com Subject: Re: [time-nuts] TAPR TICC and TADD-2 Minis On 6/1/20 9:11 AM, Bob kb8tq wrote: > Scope basics: Thanks for the comments. Very helpful as background beyond this immediate project. > A 10 db pad ahead of the feed through is another way to deal with this. It should at > lest get you into the 20 db return loss range. I have a nice JFW variable attenuator, and inserted it between the TADD-2 Mini and the scope to capture the Thunderbolt's 10 MHz signal converted to square wave with 0, 3, 6, 10 and 20 dB attenuation. Screen shots are attached (very small, but there are five of them). > One of the reasons there are compensated scope probes is to get around this > issue. There are a variety of them out there with a range of capabilities. Yes, the > fancy fet probes cost more than something simple ……. Currently the only probes I have is the stock pair that came with the Siglent digital scope (an SDS1202?X-e). I also recently bought a vintage Tektronix T922 analog scope, but it didn't include probes. Thus I might be in the market for a good set of probes. If there are suggestions for any that aren't more expensive than, say, a T922 or a used Thunderbolt, I'd be very interested. Thanks again, Frank
J
jimlux
Tue, Jun 2, 2020 9:53 PM

On 6/2/20 2:06 PM, Tom Holmes wrote:

Frank...

What is interesting is that all of your traces appear to be identical except for the vertical scale factor, which suggests that the ringing you are seeing, which is much improved from your earlier plots, is still a little bit of a problem. I am going to guess that , as Bob pointed out, is probably due to a reflection problem in the input side of the attenuator. Could be the attenuator input impedance is not a good match to the source, or could be a cable impedance issue.

I hope that suggests some things you could try experimentally and educationally to sort it out.

One trick is to move the attenuator to the other end of the cable (i.e.
at the output of the device) and then run the coax to your 50 ohm scope
input, and see what it looks like.

On 6/2/20 2:06 PM, Tom Holmes wrote: > Frank... > > What is interesting is that all of your traces appear to be identical except for the vertical scale factor, which suggests that the ringing you are seeing, which is much improved from your earlier plots, is still a little bit of a problem. I am going to guess that , as Bob pointed out, is probably due to a reflection problem in the input side of the attenuator. Could be the attenuator input impedance is not a good match to the source, or could be a cable impedance issue. > > I hope that suggests some things you could try experimentally and educationally to sort it out. > One trick is to move the attenuator to the other end of the cable (i.e. at the output of the device) and then run the coax to your 50 ohm scope input, and see what it looks like.