Unless you have a way to know what they are doing. You won't know if they
are failing. The initial sign is that they wink-out and restart. You will
spot it as your gear going nuts for a bit and then magically locking up.
Also as an approx lamp V in the 2-3V range. Thats really a gross figure I
can't swear thats EOL. But its some place close to that.
As such I installed the Rb winkOmeter on my unit.
Regards
Paul
WB8TSL
On Thu, Nov 20, 2014 at 6:09 PM, Bob Camp kb8tq@n1k.org wrote:
Hi
So far, I have yet to have one of the Rb’s die (of the ones that worked
when new to me). The magic “lamp voltage” on the Efratom versions clusters
in two groups. That seems to be related to some sort of production change
rather than better performance on the (slightly) newer parts.
Bob
On Nov 20, 2014, at 9:17 AM, paul swed paulswedb@gmail.com wrote:
On the old telco RBs $25 they all worked, but were down on lamp life. So
the first one chosen ran for 3 years and then started blinking out.
Replaced it with another. That first one is the one I used the heat gun
and is again operable with what appears could be good lamp life.
(Reasonable)
Regards
Paul
WB8TSL
On Thu, Nov 20, 2014 at 7:00 AM, Bob Camp kb8tq@n1k.org wrote:
Hi
So far, I have not found any of the Rb’s that came out of the KS boxes
with troubles. The ones I got and pulled from boxes myself all have
fine. The units that got pulled on the other side of an ocean, not so
There have been a couple of them with issues. What ever they go through
getting shipped overseas and salvaged, it’s not doing the parts any
That’s one of the reasons getting the newer design KS boxes NIB has a
of appeal. There are infant mortality reliability issues to deal with on
new stuff. I’d rather have those issues than end of life and corrosion.
Bob
On Nov 19, 2014, at 9:05 PM, paul swed paulswedb@gmail.com wrote:
And I am the happy owner of 4 or 5 of the $25 RBs. Hey they are old but
can recover them with the old heatgun trick. That said glowing things
those do tend to wear out. I could see why they moved away from them.
On Wed, Nov 19, 2014 at 8:15 PM, Bob Camp kb8tq@n1k.org wrote:
Hi
Lucent has been putting GPS based Radio Frequency Timing Generators
into cell sites for a long time. They all have the 15 MHz and pps
in common. Lucent also seems to like one paint color and an
size. For what ever reason (because they can?) the model numbers and
internal designs change from time to time. For quite a while they
liked the idea of running Rb’s in cell sites. Looking at the latest
of parts, that seems to have changed around 1999.
Lots of generations of very similar looking boxes with very different
stuff inside each of them.
Bob
On Nov 19, 2014, at 7:53 PM, paul swed paulswedb@gmail.com wrote:
I just saw the picture of the units front panel and it looked the
Oh well.
Regards
Paul
WB8TSL
On Wed, Nov 19, 2014 at 7:08 PM, Bob Camp kb8tq@n1k.org wrote:
Hi
This is a branch thread that never got it’s title changed to drop
KS-24361 part of it. More or less it’s “compare the older KS boxes
On Nov 19, 2014, at 6:03 PM, Dan Rae danrae@verizon.net wrote:
On 11/19/2014 2:47 PM, paul swed wrote:
Was doing a search for Lucent RFTGm-II-XO looking for alternates
Paul, that has nothing to do with the current ebay offerings that
generated so much traffic recently...
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
Unless you have a way to know what they are doing. You won't know if they
are failing. The initial sign is that they wink-out and restart. You will
spot it as your gear going nuts for a bit and then magically locking up.
Also as an approx lamp V in the 2-3V range. Thats really a gross figure I
can't swear thats EOL. But its some place close to that.
As such I installed the Rb winkOmeter on my unit.
Regards
Paul
WB8TSL
On Thu, Nov 20, 2014 at 6:09 PM, Bob Camp <kb8tq@n1k.org> wrote:
> Hi
>
> So far, I have yet to have one of the Rb’s die (of the ones that worked
> when new to me). The magic “lamp voltage” on the Efratom versions clusters
> in two groups. That seems to be related to some sort of production change
> rather than better performance on the (slightly) newer parts.
>
> Bob
>
> > On Nov 20, 2014, at 9:17 AM, paul swed <paulswedb@gmail.com> wrote:
> >
> > On the old telco RBs $25 they all worked, but were down on lamp life. So
> > the first one chosen ran for 3 years and then started blinking out.
> > Replaced it with another. That first one is the one I used the heat gun
> on
> > and is again operable with what appears could be good lamp life.
> > (Reasonable)
> > Regards
> > Paul
> > WB8TSL
> >
> > On Thu, Nov 20, 2014 at 7:00 AM, Bob Camp <kb8tq@n1k.org> wrote:
> >
> >> Hi
> >>
> >> So far, I have not found any of the Rb’s that came out of the KS boxes
> >> with troubles. The ones I got and pulled from boxes myself all have
> worked
> >> fine. The units that got pulled on the other side of an ocean, not so
> much.
> >> There have been a couple of them with issues. What ever they go through
> >> getting shipped overseas and salvaged, it’s not doing the parts any
> good.
> >> That’s one of the reasons getting the newer design KS boxes NIB has a
> lot
> >> of appeal. There are infant mortality reliability issues to deal with on
> >> new stuff. I’d rather have those issues than end of life and corrosion.
> >>
> >> Bob
> >>
> >>
> >>> On Nov 19, 2014, at 9:05 PM, paul swed <paulswedb@gmail.com> wrote:
> >>>
> >>> And I am the happy owner of 4 or 5 of the $25 RBs. Hey they are old but
> >> you
> >>> can recover them with the old heatgun trick. That said glowing things
> >> like
> >>> those do tend to wear out. I could see why they moved away from them.
> >>>
> >>> On Wed, Nov 19, 2014 at 8:15 PM, Bob Camp <kb8tq@n1k.org> wrote:
> >>>
> >>>> Hi
> >>>>
> >>>> Lucent has been putting GPS based Radio Frequency Timing Generators
> >> (RFTG)
> >>>> into cell sites for a long time. They all have the 15 MHz and pps
> >> outputs
> >>>> in common. Lucent also seems to like one paint color and an
> approximate
> >>>> size. For what ever reason (because they can?) the model numbers and
> >>>> internal designs change from time to time. For quite a while they
> really
> >>>> liked the idea of running Rb’s in cell sites. Looking at the latest
> >> batch
> >>>> of parts, that seems to have changed around 1999.
> >>>>
> >>>> Lots of generations of very similar looking boxes with very different
> >>>> stuff inside each of them.
> >>>>
> >>>> Bob
> >>>>
> >>>>> On Nov 19, 2014, at 7:53 PM, paul swed <paulswedb@gmail.com> wrote:
> >>>>>
> >>>>> I just saw the picture of the units front panel and it looked the
> same.
> >>>>> Oh well.
> >>>>> Regards
> >>>>> Paul
> >>>>> WB8TSL
> >>>>>
> >>>>> On Wed, Nov 19, 2014 at 7:08 PM, Bob Camp <kb8tq@n1k.org> wrote:
> >>>>>
> >>>>>> Hi
> >>>>>>
> >>>>>> This is a branch thread that never got it’s title changed to drop
> the
> >>>>>> KS-24361 part of it. More or less it’s “compare the older KS boxes
> to
> >>>> the
> >>>>>> newer ones”.
> >>>>>>
> >>>>>> Bob
> >>>>>>
> >>>>>>> On Nov 19, 2014, at 6:03 PM, Dan Rae <danrae@verizon.net> wrote:
> >>>>>>>
> >>>>>>> On 11/19/2014 2:47 PM, paul swed wrote:
> >>>>>>>> Was doing a search for Lucent RFTGm-II-XO looking for alternates
> to
> >>>> the
> >>>>>>>> $150 solution.
> >>>>>>>> A 10 Mhz mod popped up.
> >>>>>>>>
> >>>>>>>> http://telcodata.us/~myself/Lucent-RFTGm-Modification.pdf
> >>>>>>>>
> >>>>>>>> Looks simple enough and I have not validated anything.
> >>>>>>>> Regards
> >>>>>>>> Paul
> >>>>>>>> WB8TSL
> >>>>>>>>
> >>>>>>> Paul, that has nothing to do with the current ebay offerings that
> >> have
> >>>>>> generated so much traffic recently...
> >>>>>>>
> >>>>>>> Dan
> >>>>>>> _______________________________________________
> >>>>>>> time-nuts mailing list -- time-nuts@febo.com
> >>>>>>> To unsubscribe, go to
> >>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> >>>>>>> and follow the instructions there.
> >>>>>>
> >>>>>> _______________________________________________
> >>>>>> time-nuts mailing list -- time-nuts@febo.com
> >>>>>> To unsubscribe, go to
> >>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> >>>>>> and follow the instructions there.
> >>>>>>
> >>>>> _______________________________________________
> >>>>> time-nuts mailing list -- time-nuts@febo.com
> >>>>> To unsubscribe, go to
> >>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> >>>>> and follow the instructions there.
> >>>>
> >>>> _______________________________________________
> >>>> time-nuts mailing list -- time-nuts@febo.com
> >>>> To unsubscribe, go to
> >>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> >>>> and follow the instructions there.
> >>>>
> >>> _______________________________________________
> >>> time-nuts mailing list -- time-nuts@febo.com
> >>> To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> >>> and follow the instructions there.
> >>
> >> _______________________________________________
> >> time-nuts mailing list -- time-nuts@febo.com
> >> To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> >> and follow the instructions there.
> >>
> > _______________________________________________
> > time-nuts mailing list -- time-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> > and follow the instructions there.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
BC
Bob Camp
Fri, Nov 21, 2014 2:07 AM
Hi
I normally keep a watch on the lamp voltage (or what ever it actually is). I was once told by somebody who should know that you simply watch for it to drop. When it starts going down a couple volts, that’s when you should start looking for a backup. It may run for months after it starts to drop. It may run for weeks. There apparently is no good way to be sure. On a part that starts at 6 to 7 volts, that would equate to something a 3 to 4 volt limit.
Bob
On Nov 20, 2014, at 9:00 PM, paul swed paulswedb@gmail.com wrote:
Unless you have a way to know what they are doing. You won't know if they
are failing. The initial sign is that they wink-out and restart. You will
spot it as your gear going nuts for a bit and then magically locking up.
Also as an approx lamp V in the 2-3V range. Thats really a gross figure I
can't swear thats EOL. But its some place close to that.
As such I installed the Rb winkOmeter on my unit.
Regards
Paul
WB8TSL
On Thu, Nov 20, 2014 at 6:09 PM, Bob Camp kb8tq@n1k.org wrote:
Hi
So far, I have yet to have one of the Rb’s die (of the ones that worked
when new to me). The magic “lamp voltage” on the Efratom versions clusters
in two groups. That seems to be related to some sort of production change
rather than better performance on the (slightly) newer parts.
Bob
On Nov 20, 2014, at 9:17 AM, paul swed paulswedb@gmail.com wrote:
On the old telco RBs $25 they all worked, but were down on lamp life. So
the first one chosen ran for 3 years and then started blinking out.
Replaced it with another. That first one is the one I used the heat gun
and is again operable with what appears could be good lamp life.
(Reasonable)
Regards
Paul
WB8TSL
On Thu, Nov 20, 2014 at 7:00 AM, Bob Camp kb8tq@n1k.org wrote:
Hi
So far, I have not found any of the Rb’s that came out of the KS boxes
with troubles. The ones I got and pulled from boxes myself all have
fine. The units that got pulled on the other side of an ocean, not so
There have been a couple of them with issues. What ever they go through
getting shipped overseas and salvaged, it’s not doing the parts any
That’s one of the reasons getting the newer design KS boxes NIB has a
of appeal. There are infant mortality reliability issues to deal with on
new stuff. I’d rather have those issues than end of life and corrosion.
Bob
On Nov 19, 2014, at 9:05 PM, paul swed paulswedb@gmail.com wrote:
And I am the happy owner of 4 or 5 of the $25 RBs. Hey they are old but
can recover them with the old heatgun trick. That said glowing things
those do tend to wear out. I could see why they moved away from them.
On Wed, Nov 19, 2014 at 8:15 PM, Bob Camp kb8tq@n1k.org wrote:
Hi
Lucent has been putting GPS based Radio Frequency Timing Generators
into cell sites for a long time. They all have the 15 MHz and pps
in common. Lucent also seems to like one paint color and an
size. For what ever reason (because they can?) the model numbers and
internal designs change from time to time. For quite a while they
liked the idea of running Rb’s in cell sites. Looking at the latest
of parts, that seems to have changed around 1999.
Lots of generations of very similar looking boxes with very different
stuff inside each of them.
Bob
On Nov 19, 2014, at 7:53 PM, paul swed paulswedb@gmail.com wrote:
I just saw the picture of the units front panel and it looked the
Oh well.
Regards
Paul
WB8TSL
On Wed, Nov 19, 2014 at 7:08 PM, Bob Camp kb8tq@n1k.org wrote:
Hi
This is a branch thread that never got it’s title changed to drop
KS-24361 part of it. More or less it’s “compare the older KS boxes
On Nov 19, 2014, at 6:03 PM, Dan Rae danrae@verizon.net wrote:
On 11/19/2014 2:47 PM, paul swed wrote:
Was doing a search for Lucent RFTGm-II-XO looking for alternates
Paul, that has nothing to do with the current ebay offerings that
generated so much traffic recently...
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
Hi
I normally keep a watch on the lamp voltage (or what ever it actually is). I was once told by somebody who should know that you simply watch for it to drop. When it starts going down a couple volts, that’s when you should start looking for a backup. It may run for months after it starts to drop. It may run for weeks. There apparently is no good way to be sure. On a part that starts at 6 to 7 volts, that would equate to something a 3 to 4 volt limit.
Bob
> On Nov 20, 2014, at 9:00 PM, paul swed <paulswedb@gmail.com> wrote:
>
> Unless you have a way to know what they are doing. You won't know if they
> are failing. The initial sign is that they wink-out and restart. You will
> spot it as your gear going nuts for a bit and then magically locking up.
> Also as an approx lamp V in the 2-3V range. Thats really a gross figure I
> can't swear thats EOL. But its some place close to that.
> As such I installed the Rb winkOmeter on my unit.
> Regards
> Paul
> WB8TSL
>
> On Thu, Nov 20, 2014 at 6:09 PM, Bob Camp <kb8tq@n1k.org> wrote:
>
>> Hi
>>
>> So far, I have yet to have one of the Rb’s die (of the ones that worked
>> when new to me). The magic “lamp voltage” on the Efratom versions clusters
>> in two groups. That seems to be related to some sort of production change
>> rather than better performance on the (slightly) newer parts.
>>
>> Bob
>>
>>> On Nov 20, 2014, at 9:17 AM, paul swed <paulswedb@gmail.com> wrote:
>>>
>>> On the old telco RBs $25 they all worked, but were down on lamp life. So
>>> the first one chosen ran for 3 years and then started blinking out.
>>> Replaced it with another. That first one is the one I used the heat gun
>> on
>>> and is again operable with what appears could be good lamp life.
>>> (Reasonable)
>>> Regards
>>> Paul
>>> WB8TSL
>>>
>>> On Thu, Nov 20, 2014 at 7:00 AM, Bob Camp <kb8tq@n1k.org> wrote:
>>>
>>>> Hi
>>>>
>>>> So far, I have not found any of the Rb’s that came out of the KS boxes
>>>> with troubles. The ones I got and pulled from boxes myself all have
>> worked
>>>> fine. The units that got pulled on the other side of an ocean, not so
>> much.
>>>> There have been a couple of them with issues. What ever they go through
>>>> getting shipped overseas and salvaged, it’s not doing the parts any
>> good.
>>>> That’s one of the reasons getting the newer design KS boxes NIB has a
>> lot
>>>> of appeal. There are infant mortality reliability issues to deal with on
>>>> new stuff. I’d rather have those issues than end of life and corrosion.
>>>>
>>>> Bob
>>>>
>>>>
>>>>> On Nov 19, 2014, at 9:05 PM, paul swed <paulswedb@gmail.com> wrote:
>>>>>
>>>>> And I am the happy owner of 4 or 5 of the $25 RBs. Hey they are old but
>>>> you
>>>>> can recover them with the old heatgun trick. That said glowing things
>>>> like
>>>>> those do tend to wear out. I could see why they moved away from them.
>>>>>
>>>>> On Wed, Nov 19, 2014 at 8:15 PM, Bob Camp <kb8tq@n1k.org> wrote:
>>>>>
>>>>>> Hi
>>>>>>
>>>>>> Lucent has been putting GPS based Radio Frequency Timing Generators
>>>> (RFTG)
>>>>>> into cell sites for a long time. They all have the 15 MHz and pps
>>>> outputs
>>>>>> in common. Lucent also seems to like one paint color and an
>> approximate
>>>>>> size. For what ever reason (because they can?) the model numbers and
>>>>>> internal designs change from time to time. For quite a while they
>> really
>>>>>> liked the idea of running Rb’s in cell sites. Looking at the latest
>>>> batch
>>>>>> of parts, that seems to have changed around 1999.
>>>>>>
>>>>>> Lots of generations of very similar looking boxes with very different
>>>>>> stuff inside each of them.
>>>>>>
>>>>>> Bob
>>>>>>
>>>>>>> On Nov 19, 2014, at 7:53 PM, paul swed <paulswedb@gmail.com> wrote:
>>>>>>>
>>>>>>> I just saw the picture of the units front panel and it looked the
>> same.
>>>>>>> Oh well.
>>>>>>> Regards
>>>>>>> Paul
>>>>>>> WB8TSL
>>>>>>>
>>>>>>> On Wed, Nov 19, 2014 at 7:08 PM, Bob Camp <kb8tq@n1k.org> wrote:
>>>>>>>
>>>>>>>> Hi
>>>>>>>>
>>>>>>>> This is a branch thread that never got it’s title changed to drop
>> the
>>>>>>>> KS-24361 part of it. More or less it’s “compare the older KS boxes
>> to
>>>>>> the
>>>>>>>> newer ones”.
>>>>>>>>
>>>>>>>> Bob
>>>>>>>>
>>>>>>>>> On Nov 19, 2014, at 6:03 PM, Dan Rae <danrae@verizon.net> wrote:
>>>>>>>>>
>>>>>>>>> On 11/19/2014 2:47 PM, paul swed wrote:
>>>>>>>>>> Was doing a search for Lucent RFTGm-II-XO looking for alternates
>> to
>>>>>> the
>>>>>>>>>> $150 solution.
>>>>>>>>>> A 10 Mhz mod popped up.
>>>>>>>>>>
>>>>>>>>>> http://telcodata.us/~myself/Lucent-RFTGm-Modification.pdf
>>>>>>>>>>
>>>>>>>>>> Looks simple enough and I have not validated anything.
>>>>>>>>>> Regards
>>>>>>>>>> Paul
>>>>>>>>>> WB8TSL
>>>>>>>>>>
>>>>>>>>> Paul, that has nothing to do with the current ebay offerings that
>>>> have
>>>>>>>> generated so much traffic recently...
>>>>>>>>>
>>>>>>>>> Dan
>>>>>>>>> _______________________________________________
>>>>>>>>> time-nuts mailing list -- time-nuts@febo.com
>>>>>>>>> To unsubscribe, go to
>>>>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>>>>>> and follow the instructions there.
>>>>>>>>
>>>>>>>> _______________________________________________
>>>>>>>> time-nuts mailing list -- time-nuts@febo.com
>>>>>>>> To unsubscribe, go to
>>>>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>>>>> and follow the instructions there.
>>>>>>>>
>>>>>>> _______________________________________________
>>>>>>> time-nuts mailing list -- time-nuts@febo.com
>>>>>>> To unsubscribe, go to
>>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>>>> and follow the instructions there.
>>>>>>
>>>>>> _______________________________________________
>>>>>> time-nuts mailing list -- time-nuts@febo.com
>>>>>> To unsubscribe, go to
>>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>>> and follow the instructions there.
>>>>>>
>>>>> _______________________________________________
>>>>> time-nuts mailing list -- time-nuts@febo.com
>>>>> To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>> and follow the instructions there.
>>>>
>>>> _______________________________________________
>>>> time-nuts mailing list -- time-nuts@febo.com
>>>> To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>> and follow the instructions there.
>>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts@febo.com
>>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
PS
paul swed
Fri, Nov 21, 2014 3:31 AM
I do agree its subjective. But my gut says bad is 2-3V I know they seem to
run above 3. But whatever. When the blinkOmeter counts I am screwed. :-)
Time to get the heatgun.
Regards
Paul.
On Thu, Nov 20, 2014 at 9:07 PM, Bob Camp kb8tq@n1k.org wrote:
Hi
I normally keep a watch on the lamp voltage (or what ever it actually is).
I was once told by somebody who should know that you simply watch for it to
drop. When it starts going down a couple volts, that’s when you should
start looking for a backup. It may run for months after it starts to drop.
It may run for weeks. There apparently is no good way to be sure. On a part
that starts at 6 to 7 volts, that would equate to something a 3 to 4 volt
limit.
Bob
On Nov 20, 2014, at 9:00 PM, paul swed paulswedb@gmail.com wrote:
Unless you have a way to know what they are doing. You won't know if they
are failing. The initial sign is that they wink-out and restart. You will
spot it as your gear going nuts for a bit and then magically locking up.
Also as an approx lamp V in the 2-3V range. Thats really a gross figure I
can't swear thats EOL. But its some place close to that.
As such I installed the Rb winkOmeter on my unit.
Regards
Paul
WB8TSL
On Thu, Nov 20, 2014 at 6:09 PM, Bob Camp kb8tq@n1k.org wrote:
Hi
So far, I have yet to have one of the Rb’s die (of the ones that worked
when new to me). The magic “lamp voltage” on the Efratom versions
in two groups. That seems to be related to some sort of production
rather than better performance on the (slightly) newer parts.
Bob
On Nov 20, 2014, at 9:17 AM, paul swed paulswedb@gmail.com wrote:
On the old telco RBs $25 they all worked, but were down on lamp life.
the first one chosen ran for 3 years and then started blinking out.
Replaced it with another. That first one is the one I used the heat gun
and is again operable with what appears could be good lamp life.
(Reasonable)
Regards
Paul
WB8TSL
On Thu, Nov 20, 2014 at 7:00 AM, Bob Camp kb8tq@n1k.org wrote:
Hi
So far, I have not found any of the Rb’s that came out of the KS boxes
with troubles. The ones I got and pulled from boxes myself all have
fine. The units that got pulled on the other side of an ocean, not so
There have been a couple of them with issues. What ever they go
getting shipped overseas and salvaged, it’s not doing the parts any
That’s one of the reasons getting the newer design KS boxes NIB has a
of appeal. There are infant mortality reliability issues to deal with
new stuff. I’d rather have those issues than end of life and
On Nov 19, 2014, at 9:05 PM, paul swed paulswedb@gmail.com wrote:
And I am the happy owner of 4 or 5 of the $25 RBs. Hey they are old
can recover them with the old heatgun trick. That said glowing things
those do tend to wear out. I could see why they moved away from them.
On Wed, Nov 19, 2014 at 8:15 PM, Bob Camp kb8tq@n1k.org wrote:
Hi
Lucent has been putting GPS based Radio Frequency Timing Generators
into cell sites for a long time. They all have the 15 MHz and pps
in common. Lucent also seems to like one paint color and an
size. For what ever reason (because they can?) the model numbers and
internal designs change from time to time. For quite a while they
liked the idea of running Rb’s in cell sites. Looking at the latest
of parts, that seems to have changed around 1999.
Lots of generations of very similar looking boxes with very
stuff inside each of them.
Bob
I just saw the picture of the units front panel and it looked the
Oh well.
Regards
Paul
WB8TSL
On Wed, Nov 19, 2014 at 7:08 PM, Bob Camp kb8tq@n1k.org wrote:
Hi
This is a branch thread that never got it’s title changed to drop
KS-24361 part of it. More or less it’s “compare the older KS boxes
On Nov 19, 2014, at 6:03 PM, Dan Rae danrae@verizon.net wrote:
On 11/19/2014 2:47 PM, paul swed wrote:
Was doing a search for Lucent RFTGm-II-XO looking for alternates
Paul, that has nothing to do with the current ebay offerings that
generated so much traffic recently...
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
I do agree its subjective. But my gut says bad is 2-3V I know they seem to
run above 3. But whatever. When the blinkOmeter counts I am screwed. :-)
Time to get the heatgun.
Regards
Paul.
On Thu, Nov 20, 2014 at 9:07 PM, Bob Camp <kb8tq@n1k.org> wrote:
> Hi
>
> I normally keep a watch on the lamp voltage (or what ever it actually is).
> I was once told by somebody who should know that you simply watch for it to
> drop. When it starts going down a couple volts, that’s when you should
> start looking for a backup. It may run for months after it starts to drop.
> It may run for weeks. There apparently is no good way to be sure. On a part
> that starts at 6 to 7 volts, that would equate to something a 3 to 4 volt
> limit.
>
> Bob
>
> > On Nov 20, 2014, at 9:00 PM, paul swed <paulswedb@gmail.com> wrote:
> >
> > Unless you have a way to know what they are doing. You won't know if they
> > are failing. The initial sign is that they wink-out and restart. You will
> > spot it as your gear going nuts for a bit and then magically locking up.
> > Also as an approx lamp V in the 2-3V range. Thats really a gross figure I
> > can't swear thats EOL. But its some place close to that.
> > As such I installed the Rb winkOmeter on my unit.
> > Regards
> > Paul
> > WB8TSL
> >
> > On Thu, Nov 20, 2014 at 6:09 PM, Bob Camp <kb8tq@n1k.org> wrote:
> >
> >> Hi
> >>
> >> So far, I have yet to have one of the Rb’s die (of the ones that worked
> >> when new to me). The magic “lamp voltage” on the Efratom versions
> clusters
> >> in two groups. That seems to be related to some sort of production
> change
> >> rather than better performance on the (slightly) newer parts.
> >>
> >> Bob
> >>
> >>> On Nov 20, 2014, at 9:17 AM, paul swed <paulswedb@gmail.com> wrote:
> >>>
> >>> On the old telco RBs $25 they all worked, but were down on lamp life.
> So
> >>> the first one chosen ran for 3 years and then started blinking out.
> >>> Replaced it with another. That first one is the one I used the heat gun
> >> on
> >>> and is again operable with what appears could be good lamp life.
> >>> (Reasonable)
> >>> Regards
> >>> Paul
> >>> WB8TSL
> >>>
> >>> On Thu, Nov 20, 2014 at 7:00 AM, Bob Camp <kb8tq@n1k.org> wrote:
> >>>
> >>>> Hi
> >>>>
> >>>> So far, I have not found any of the Rb’s that came out of the KS boxes
> >>>> with troubles. The ones I got and pulled from boxes myself all have
> >> worked
> >>>> fine. The units that got pulled on the other side of an ocean, not so
> >> much.
> >>>> There have been a couple of them with issues. What ever they go
> through
> >>>> getting shipped overseas and salvaged, it’s not doing the parts any
> >> good.
> >>>> That’s one of the reasons getting the newer design KS boxes NIB has a
> >> lot
> >>>> of appeal. There are infant mortality reliability issues to deal with
> on
> >>>> new stuff. I’d rather have those issues than end of life and
> corrosion.
> >>>>
> >>>> Bob
> >>>>
> >>>>
> >>>>> On Nov 19, 2014, at 9:05 PM, paul swed <paulswedb@gmail.com> wrote:
> >>>>>
> >>>>> And I am the happy owner of 4 or 5 of the $25 RBs. Hey they are old
> but
> >>>> you
> >>>>> can recover them with the old heatgun trick. That said glowing things
> >>>> like
> >>>>> those do tend to wear out. I could see why they moved away from them.
> >>>>>
> >>>>> On Wed, Nov 19, 2014 at 8:15 PM, Bob Camp <kb8tq@n1k.org> wrote:
> >>>>>
> >>>>>> Hi
> >>>>>>
> >>>>>> Lucent has been putting GPS based Radio Frequency Timing Generators
> >>>> (RFTG)
> >>>>>> into cell sites for a long time. They all have the 15 MHz and pps
> >>>> outputs
> >>>>>> in common. Lucent also seems to like one paint color and an
> >> approximate
> >>>>>> size. For what ever reason (because they can?) the model numbers and
> >>>>>> internal designs change from time to time. For quite a while they
> >> really
> >>>>>> liked the idea of running Rb’s in cell sites. Looking at the latest
> >>>> batch
> >>>>>> of parts, that seems to have changed around 1999.
> >>>>>>
> >>>>>> Lots of generations of very similar looking boxes with very
> different
> >>>>>> stuff inside each of them.
> >>>>>>
> >>>>>> Bob
> >>>>>>
> >>>>>>> On Nov 19, 2014, at 7:53 PM, paul swed <paulswedb@gmail.com>
> wrote:
> >>>>>>>
> >>>>>>> I just saw the picture of the units front panel and it looked the
> >> same.
> >>>>>>> Oh well.
> >>>>>>> Regards
> >>>>>>> Paul
> >>>>>>> WB8TSL
> >>>>>>>
> >>>>>>> On Wed, Nov 19, 2014 at 7:08 PM, Bob Camp <kb8tq@n1k.org> wrote:
> >>>>>>>
> >>>>>>>> Hi
> >>>>>>>>
> >>>>>>>> This is a branch thread that never got it’s title changed to drop
> >> the
> >>>>>>>> KS-24361 part of it. More or less it’s “compare the older KS boxes
> >> to
> >>>>>> the
> >>>>>>>> newer ones”.
> >>>>>>>>
> >>>>>>>> Bob
> >>>>>>>>
> >>>>>>>>> On Nov 19, 2014, at 6:03 PM, Dan Rae <danrae@verizon.net> wrote:
> >>>>>>>>>
> >>>>>>>>> On 11/19/2014 2:47 PM, paul swed wrote:
> >>>>>>>>>> Was doing a search for Lucent RFTGm-II-XO looking for alternates
> >> to
> >>>>>> the
> >>>>>>>>>> $150 solution.
> >>>>>>>>>> A 10 Mhz mod popped up.
> >>>>>>>>>>
> >>>>>>>>>> http://telcodata.us/~myself/Lucent-RFTGm-Modification.pdf
> >>>>>>>>>>
> >>>>>>>>>> Looks simple enough and I have not validated anything.
> >>>>>>>>>> Regards
> >>>>>>>>>> Paul
> >>>>>>>>>> WB8TSL
> >>>>>>>>>>
> >>>>>>>>> Paul, that has nothing to do with the current ebay offerings that
> >>>> have
> >>>>>>>> generated so much traffic recently...
> >>>>>>>>>
> >>>>>>>>> Dan
> >>>>>>>>> _______________________________________________
> >>>>>>>>> time-nuts mailing list -- time-nuts@febo.com
> >>>>>>>>> To unsubscribe, go to
> >>>>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> >>>>>>>>> and follow the instructions there.
> >>>>>>>>
> >>>>>>>> _______________________________________________
> >>>>>>>> time-nuts mailing list -- time-nuts@febo.com
> >>>>>>>> To unsubscribe, go to
> >>>>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> >>>>>>>> and follow the instructions there.
> >>>>>>>>
> >>>>>>> _______________________________________________
> >>>>>>> time-nuts mailing list -- time-nuts@febo.com
> >>>>>>> To unsubscribe, go to
> >>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> >>>>>>> and follow the instructions there.
> >>>>>>
> >>>>>> _______________________________________________
> >>>>>> time-nuts mailing list -- time-nuts@febo.com
> >>>>>> To unsubscribe, go to
> >>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> >>>>>> and follow the instructions there.
> >>>>>>
> >>>>> _______________________________________________
> >>>>> time-nuts mailing list -- time-nuts@febo.com
> >>>>> To unsubscribe, go to
> >>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> >>>>> and follow the instructions there.
> >>>>
> >>>> _______________________________________________
> >>>> time-nuts mailing list -- time-nuts@febo.com
> >>>> To unsubscribe, go to
> >>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> >>>> and follow the instructions there.
> >>>>
> >>> _______________________________________________
> >>> time-nuts mailing list -- time-nuts@febo.com
> >>> To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> >>> and follow the instructions there.
> >>
> >> _______________________________________________
> >> time-nuts mailing list -- time-nuts@febo.com
> >> To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> >> and follow the instructions there.
> >>
> > _______________________________________________
> > time-nuts mailing list -- time-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> > and follow the instructions there.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
AP
Alex Pummer
Fri, Nov 21, 2014 3:41 AM
but there was a description; how to rejuvenate the rubidium bulb by
cautiously warning it up, to "remelt " the rubidium....
73
Alex
On 11/20/2014 7:31 PM, paul swed wrote:
I do agree its subjective. But my gut says bad is 2-3V I know they seem to
run above 3. But whatever. When the blinkOmeter counts I am screwed. :-)
Time to get the heatgun.
Regards
Paul.
On Thu, Nov 20, 2014 at 9:07 PM, Bob Camp kb8tq@n1k.org wrote:
but there was a description; how to rejuvenate the rubidium bulb by
cautiously warning it up, to "remelt " the rubidium....
73
Alex
On 11/20/2014 7:31 PM, paul swed wrote:
> I do agree its subjective. But my gut says bad is 2-3V I know they seem to
> run above 3. But whatever. When the blinkOmeter counts I am screwed. :-)
> Time to get the heatgun.
> Regards
> Paul.
>
> On Thu, Nov 20, 2014 at 9:07 PM, Bob Camp <kb8tq@n1k.org> wrote:
>
>
BC
Bob Camp
Fri, Nov 21, 2014 12:05 PM
Hi
The heat gun approach assumes that the fault is a blob of metal that is stuck in a bad location in the bulb. Unfortunately there are a lot of different ways these beasts can die. The repair guys who should know always claimed that bulb issues were < 30% of the issues they saw. That was on an earlier generation of Rb’s but I suspect it applies. None of the ones I had arrive DOA were bulb problems.
Bob
On Nov 20, 2014, at 10:41 PM, Alex Pummer alex@pcscons.com wrote:
but there was a description; how to rejuvenate the rubidium bulb by cautiously warning it up, to "remelt " the rubidium....
73
Alex
On 11/20/2014 7:31 PM, paul swed wrote:
I do agree its subjective. But my gut says bad is 2-3V I know they seem to
run above 3. But whatever. When the blinkOmeter counts I am screwed. :-)
Time to get the heatgun.
Regards
Paul.
On Thu, Nov 20, 2014 at 9:07 PM, Bob Camp kb8tq@n1k.org wrote:
Hi
The heat gun approach assumes that the fault is a blob of metal that is stuck in a bad location in the bulb. Unfortunately there are a lot of different ways these beasts can die. The repair guys who should know always claimed that bulb issues were < 30% of the issues they saw. That was on an earlier generation of Rb’s but I suspect it applies. None of the ones I had arrive DOA were bulb problems.
Bob
> On Nov 20, 2014, at 10:41 PM, Alex Pummer <alex@pcscons.com> wrote:
>
> but there was a description; how to rejuvenate the rubidium bulb by cautiously warning it up, to "remelt " the rubidium....
> 73
> Alex
>
> On 11/20/2014 7:31 PM, paul swed wrote:
>> I do agree its subjective. But my gut says bad is 2-3V I know they seem to
>> run above 3. But whatever. When the blinkOmeter counts I am screwed. :-)
>> Time to get the heatgun.
>> Regards
>> Paul.
>>
>> On Thu, Nov 20, 2014 at 9:07 PM, Bob Camp <kb8tq@n1k.org> wrote:
>>
>>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
PS
paul swed
Fri, Nov 21, 2014 7:00 PM
Bob
Absolutely lots of ways to die. Such as bad caps, bad variable cap,
resistors that are cooked..... I am pretty good with all of those and have
experienced most of them. But then comes the point of the dirty bulb and it
doesn't have to be the blob. Though you will find the blob most likely with
the dirty bulb.
The darkened window cause a loss in signal and ultimately loss of lock.
The heat gun clears both up.
Regards
Paul
WB8TSL
On Thu, Nov 20, 2014 at 10:41 PM, Alex Pummer alex@pcscons.com wrote:
but there was a description; how to rejuvenate the rubidium bulb by
cautiously warning it up, to "remelt " the rubidium....
73
Alex
On 11/20/2014 7:31 PM, paul swed wrote:
I do agree its subjective. But my gut says bad is 2-3V I know they seem to
run above 3. But whatever. When the blinkOmeter counts I am screwed. :-)
Time to get the heatgun.
Regards
Paul.
On Thu, Nov 20, 2014 at 9:07 PM, Bob Camp kb8tq@n1k.org wrote:
Bob
Absolutely lots of ways to die. Such as bad caps, bad variable cap,
resistors that are cooked..... I am pretty good with all of those and have
experienced most of them. But then comes the point of the dirty bulb and it
doesn't have to be the blob. Though you will find the blob most likely with
the dirty bulb.
The darkened window cause a loss in signal and ultimately loss of lock.
The heat gun clears both up.
Regards
Paul
WB8TSL
On Thu, Nov 20, 2014 at 10:41 PM, Alex Pummer <alex@pcscons.com> wrote:
> but there was a description; how to rejuvenate the rubidium bulb by
> cautiously warning it up, to "remelt " the rubidium....
> 73
> Alex
>
> On 11/20/2014 7:31 PM, paul swed wrote:
>
>> I do agree its subjective. But my gut says bad is 2-3V I know they seem to
>> run above 3. But whatever. When the blinkOmeter counts I am screwed. :-)
>> Time to get the heatgun.
>> Regards
>> Paul.
>>
>> On Thu, Nov 20, 2014 at 9:07 PM, Bob Camp <kb8tq@n1k.org> wrote:
>>
>>
>>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/
> mailman/listinfo/time-nuts
> and follow the instructions there.
>
BC
Bob Camp
Fri, Nov 21, 2014 11:29 PM
Hi
Tearing the physics package apart and putting it back together again can be a bit of a hassle on the small Rb’s …
Bob
On Nov 21, 2014, at 2:00 PM, paul swed paulswedb@gmail.com wrote:
Bob
Absolutely lots of ways to die. Such as bad caps, bad variable cap,
resistors that are cooked..... I am pretty good with all of those and have
experienced most of them. But then comes the point of the dirty bulb and it
doesn't have to be the blob. Though you will find the blob most likely with
the dirty bulb.
The darkened window cause a loss in signal and ultimately loss of lock.
The heat gun clears both up.
Regards
Paul
WB8TSL
On Thu, Nov 20, 2014 at 10:41 PM, Alex Pummer alex@pcscons.com wrote:
but there was a description; how to rejuvenate the rubidium bulb by
cautiously warning it up, to "remelt " the rubidium....
73
Alex
On 11/20/2014 7:31 PM, paul swed wrote:
I do agree its subjective. But my gut says bad is 2-3V I know they seem to
run above 3. But whatever. When the blinkOmeter counts I am screwed. :-)
Time to get the heatgun.
Regards
Paul.
On Thu, Nov 20, 2014 at 9:07 PM, Bob Camp kb8tq@n1k.org wrote:
Hi
Tearing the physics package apart and putting it back together again can be a bit of a hassle on the small Rb’s …
Bob
> On Nov 21, 2014, at 2:00 PM, paul swed <paulswedb@gmail.com> wrote:
>
> Bob
> Absolutely lots of ways to die. Such as bad caps, bad variable cap,
> resistors that are cooked..... I am pretty good with all of those and have
> experienced most of them. But then comes the point of the dirty bulb and it
> doesn't have to be the blob. Though you will find the blob most likely with
> the dirty bulb.
> The darkened window cause a loss in signal and ultimately loss of lock.
> The heat gun clears both up.
> Regards
> Paul
> WB8TSL
>
> On Thu, Nov 20, 2014 at 10:41 PM, Alex Pummer <alex@pcscons.com> wrote:
>
>> but there was a description; how to rejuvenate the rubidium bulb by
>> cautiously warning it up, to "remelt " the rubidium....
>> 73
>> Alex
>>
>> On 11/20/2014 7:31 PM, paul swed wrote:
>>
>>> I do agree its subjective. But my gut says bad is 2-3V I know they seem to
>>> run above 3. But whatever. When the blinkOmeter counts I am screwed. :-)
>>> Time to get the heatgun.
>>> Regards
>>> Paul.
>>>
>>> On Thu, Nov 20, 2014 at 9:07 PM, Bob Camp <kb8tq@n1k.org> wrote:
>>>
>>>
>>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/
>> mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.