Hi All,

This is my first time on the list so apologies if I am going over an old
topic.

I am repairing a HP 8662A Synthesised Signal Generator that I purchased
from a government auction. The unit was purchased without any
information about its working state and I have fixing various issues
with it. Firstly there was a digital logic problem with the keyboard
card that was causing the generator to periodically lock up. When this
was fixed the unit came to life but the output signal level was down 5dB
and this was traced to a bad step attenuator. I attempted a fix but
ended up needing to replace this to get the output steps correct.

I have now turned my attention to calibrating the reference timebase and
unfortunately this seems to be well out of spec.

The timebase is a HP 10544B 10MHz OCXO. The 10MHz output of the
oscillator seems to be ~40Hz low. I have attempted to adjust the unit
via the multi-turn capacitor adjustment but this seems to be at the end
stop and the frequency is still low. So my questions are:

  1. I assume this module is faulty as it is so far out of spec? The
    signal generator has been powered off for multiple months but even after
    having the unit powered on for 6 hours the frequency has not changed
    significantly.

  2. How repairable are these modules? I ended up blowing $200 on the
    replacement step attenuator so will need to save up for any new parts. I
    can always use an external 10MHz reference but it is always nice to have
    equipment complete.

  3. From searching the list it seems that the oven could be what is
    faulty. Would a cold over account for this much error? If so I will pull
    the module and check this.

  4. Finally is the OCXO just worn out and has aged beyond the point it
    can be adjusted back into spec?

Regards,

Denis

Hi All, This is my first time on the list so apologies if I am going over an old topic. I am repairing a HP 8662A Synthesised Signal Generator that I purchased from a government auction. The unit was purchased without any information about its working state and I have fixing various issues with it. Firstly there was a digital logic problem with the keyboard card that was causing the generator to periodically lock up. When this was fixed the unit came to life but the output signal level was down 5dB and this was traced to a bad step attenuator. I attempted a fix but ended up needing to replace this to get the output steps correct. I have now turned my attention to calibrating the reference timebase and unfortunately this seems to be well out of spec. The timebase is a HP 10544B 10MHz OCXO. The 10MHz output of the oscillator seems to be ~40Hz low. I have attempted to adjust the unit via the multi-turn capacitor adjustment but this seems to be at the end stop and the frequency is still low. So my questions are: 1) I assume this module is faulty as it is so far out of spec? The signal generator has been powered off for multiple months but even after having the unit powered on for 6 hours the frequency has not changed significantly. 2) How repairable are these modules? I ended up blowing $200 on the replacement step attenuator so will need to save up for any new parts. I can always use an external 10MHz reference but it is always nice to have equipment complete. 3) From searching the list it seems that the oven could be what is faulty. Would a cold over account for this much error? If so I will pull the module and check this. 4) Finally is the OCXO just worn out and has aged beyond the point it can be adjusted back into spec? Regards, Denis
DG
David G. McGaw
Tue, Mar 28, 2023 1:54 AM

It does sound like the oven on the 10544 has failed.  There is a lot of
info on the unit at http://leapsecond.com/museum/10544/.

Also, if you choose not to or cannot repair it, any commonly available
version 10811 oscillator is plug-and-play (only the 10811B is not).  It
will replace a 10544 with no re-wiring.

David N1HAC

On 3/26/23 6:56 PM, Denis Dowling via time-nuts wrote:

Hi All,

This is my first time on the list so apologies if I am going over an
old topic.

I am repairing a HP 8662A Synthesised Signal Generator that I
purchased from a government auction. The unit was purchased without
any information about its working state and I have fixing various
issues with it. Firstly there was a digital logic problem with the
keyboard card that was causing the generator to periodically lock up.
When this was fixed the unit came to life but the output signal level
was down 5dB and this was traced to a bad step attenuator. I attempted
a fix but ended up needing to replace this to get the output steps
correct.

I have now turned my attention to calibrating the reference timebase
and unfortunately this seems to be well out of spec.

The timebase is a HP 10544B 10MHz OCXO. The 10MHz output of the
oscillator seems to be ~40Hz low. I have attempted to adjust the unit
via the multi-turn capacitor adjustment but this seems to be at the
end stop and the frequency is still low. So my questions are:

  1. I assume this module is faulty as it is so far out of spec? The
    signal generator has been powered off for multiple months but even
    after having the unit powered on for 6 hours the frequency has not
    changed significantly.

  2. How repairable are these modules? I ended up blowing $200 on the
    replacement step attenuator so will need to save up for any new parts.
    I can always use an external 10MHz reference but it is always nice to
    have equipment complete.

  3. From searching the list it seems that the oven could be what is
    faulty. Would a cold over account for this much error? If so I will
    pull the module and check this.

  4. Finally is the OCXO just worn out and has aged beyond the point it
    can be adjusted back into spec?

Regards,

Denis


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

It does sound like the oven on the 10544 has failed.  There is a lot of info on the unit at http://leapsecond.com/museum/10544/. Also, if you choose not to or cannot repair it, any commonly available version 10811 oscillator is plug-and-play (only the 10811B is not).  It will replace a 10544 with no re-wiring. David N1HAC On 3/26/23 6:56 PM, Denis Dowling via time-nuts wrote: > Hi All, > > This is my first time on the list so apologies if I am going over an > old topic. > > I am repairing a HP 8662A Synthesised Signal Generator that I > purchased from a government auction. The unit was purchased without > any information about its working state and I have fixing various > issues with it. Firstly there was a digital logic problem with the > keyboard card that was causing the generator to periodically lock up. > When this was fixed the unit came to life but the output signal level > was down 5dB and this was traced to a bad step attenuator. I attempted > a fix but ended up needing to replace this to get the output steps > correct. > > I have now turned my attention to calibrating the reference timebase > and unfortunately this seems to be well out of spec. > > The timebase is a HP 10544B 10MHz OCXO. The 10MHz output of the > oscillator seems to be ~40Hz low. I have attempted to adjust the unit > via the multi-turn capacitor adjustment but this seems to be at the > end stop and the frequency is still low. So my questions are: > > 1) I assume this module is faulty as it is so far out of spec? The > signal generator has been powered off for multiple months but even > after having the unit powered on for 6 hours the frequency has not > changed significantly. > > 2) How repairable are these modules? I ended up blowing $200 on the > replacement step attenuator so will need to save up for any new parts. > I can always use an external 10MHz reference but it is always nice to > have equipment complete. > > 3) From searching the list it seems that the oven could be what is > faulty. Would a cold over account for this much error? If so I will > pull the module and check this. > > 4) Finally is the OCXO just worn out and has aged beyond the point it > can be adjusted back into spec? > > > Regards, > > Denis > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com
BG
Bruce Griffiths
Tue, Mar 28, 2023 6:53 AM

N.B. there's at least one error in the circuit schematic for the 10544 at leapsecond.com.
The oven monitor series resistor should be connected to the driven end of the heater winding not to the oven supply via the thermal fuse.

The corrected oven monitor connection is then consistent with the datasheet and the 10544 I used in the early 70's which was easily repaired after its RF output was inadvertently shorted.

Bruce

On 28/03/2023 14:54 NZDT David G. McGaw via time-nuts time-nuts@lists.febo.com wrote:

It does sound like the oven on the 10544 has failed.  There is a lot of
info on the unit at http://leapsecond.com/museum/10544/.

Also, if you choose not to or cannot repair it, any commonly available
version 10811 oscillator is plug-and-play (only the 10811B is not).  It
will replace a 10544 with no re-wiring.

David N1HAC

On 3/26/23 6:56 PM, Denis Dowling via time-nuts wrote:

Hi All,

This is my first time on the list so apologies if I am going over an
old topic.

I am repairing a HP 8662A Synthesised Signal Generator that I
purchased from a government auction. The unit was purchased without
any information about its working state and I have fixing various
issues with it. Firstly there was a digital logic problem with the
keyboard card that was causing the generator to periodically lock up.
When this was fixed the unit came to life but the output signal level
was down 5dB and this was traced to a bad step attenuator. I attempted
a fix but ended up needing to replace this to get the output steps
correct.

I have now turned my attention to calibrating the reference timebase
and unfortunately this seems to be well out of spec.

The timebase is a HP 10544B 10MHz OCXO. The 10MHz output of the
oscillator seems to be ~40Hz low. I have attempted to adjust the unit
via the multi-turn capacitor adjustment but this seems to be at the
end stop and the frequency is still low. So my questions are:

  1. I assume this module is faulty as it is so far out of spec? The
    signal generator has been powered off for multiple months but even
    after having the unit powered on for 6 hours the frequency has not
    changed significantly.

  2. How repairable are these modules? I ended up blowing $200 on the
    replacement step attenuator so will need to save up for any new parts.
    I can always use an external 10MHz reference but it is always nice to
    have equipment complete.

  3. From searching the list it seems that the oven could be what is
    faulty. Would a cold over account for this much error? If so I will
    pull the module and check this.

  4. Finally is the OCXO just worn out and has aged beyond the point it
    can be adjusted back into spec?

Regards,

Denis


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

N.B. there's at least one error in the circuit schematic for the 10544 at leapsecond.com. The oven monitor series resistor should be connected to the driven end of the heater winding not to the oven supply via the thermal fuse. The corrected oven monitor connection is then consistent with the datasheet and the 10544 I used in the early 70's which was easily repaired after its RF output was inadvertently shorted. Bruce > On 28/03/2023 14:54 NZDT David G. McGaw via time-nuts <time-nuts@lists.febo.com> wrote: > > > It does sound like the oven on the 10544 has failed.  There is a lot of > info on the unit at http://leapsecond.com/museum/10544/. > > Also, if you choose not to or cannot repair it, any commonly available > version 10811 oscillator is plug-and-play (only the 10811B is not).  It > will replace a 10544 with no re-wiring. > > David N1HAC > > On 3/26/23 6:56 PM, Denis Dowling via time-nuts wrote: > > Hi All, > > > > This is my first time on the list so apologies if I am going over an > > old topic. > > > > I am repairing a HP 8662A Synthesised Signal Generator that I > > purchased from a government auction. The unit was purchased without > > any information about its working state and I have fixing various > > issues with it. Firstly there was a digital logic problem with the > > keyboard card that was causing the generator to periodically lock up. > > When this was fixed the unit came to life but the output signal level > > was down 5dB and this was traced to a bad step attenuator. I attempted > > a fix but ended up needing to replace this to get the output steps > > correct. > > > > I have now turned my attention to calibrating the reference timebase > > and unfortunately this seems to be well out of spec. > > > > The timebase is a HP 10544B 10MHz OCXO. The 10MHz output of the > > oscillator seems to be ~40Hz low. I have attempted to adjust the unit > > via the multi-turn capacitor adjustment but this seems to be at the > > end stop and the frequency is still low. So my questions are: > > > > 1) I assume this module is faulty as it is so far out of spec? The > > signal generator has been powered off for multiple months but even > > after having the unit powered on for 6 hours the frequency has not > > changed significantly. > > > > 2) How repairable are these modules? I ended up blowing $200 on the > > replacement step attenuator so will need to save up for any new parts. > > I can always use an external 10MHz reference but it is always nice to > > have equipment complete. > > > > 3) From searching the list it seems that the oven could be what is > > faulty. Would a cold over account for this much error? If so I will > > pull the module and check this. > > > > 4) Finally is the OCXO just worn out and has aged beyond the point it > > can be adjusted back into spec? > > > > > > Regards, > > > > Denis > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@lists.febo.com > > To unsubscribe send an email to time-nuts-leave@lists.febo.com > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com
BC
Bob Camp
Tue, Mar 28, 2023 12:41 PM

Hi

There’s a range of things that could be wrong. Some of them
are repairable.

If you spend some “quality time” shopping, a replacement can be
had for < $50 delivered.

Bob

On Mar 26, 2023, at 6:56 PM, Denis Dowling via time-nuts time-nuts@lists.febo.com wrote:

Hi All,

This is my first time on the list so apologies if I am going over an old topic.

I am repairing a HP 8662A Synthesised Signal Generator that I purchased from a government auction. The unit was purchased without any information about its working state and I have fixing various issues with it. Firstly there was a digital logic problem with the keyboard card that was causing the generator to periodically lock up. When this was fixed the unit came to life but the output signal level was down 5dB and this was traced to a bad step attenuator. I attempted a fix but ended up needing to replace this to get the output steps correct.

I have now turned my attention to calibrating the reference timebase and unfortunately this seems to be well out of spec.

The timebase is a HP 10544B 10MHz OCXO. The 10MHz output of the oscillator seems to be ~40Hz low. I have attempted to adjust the unit via the multi-turn capacitor adjustment but this seems to be at the end stop and the frequency is still low. So my questions are:

  1. I assume this module is faulty as it is so far out of spec? The signal generator has been powered off for multiple months but even after having the unit powered on for 6 hours the frequency has not changed significantly.

  2. How repairable are these modules? I ended up blowing $200 on the replacement step attenuator so will need to save up for any new parts. I can always use an external 10MHz reference but it is always nice to have equipment complete.

  3. From searching the list it seems that the oven could be what is faulty. Would a cold over account for this much error? If so I will pull the module and check this.

  4. Finally is the OCXO just worn out and has aged beyond the point it can be adjusted back into spec?

Regards,

Denis


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

Hi There’s a range of things that *could* be wrong. Some of them are repairable. If you spend some “quality time” shopping, a replacement can be had for < $50 delivered. Bob > On Mar 26, 2023, at 6:56 PM, Denis Dowling via time-nuts <time-nuts@lists.febo.com> wrote: > > Hi All, > > This is my first time on the list so apologies if I am going over an old topic. > > I am repairing a HP 8662A Synthesised Signal Generator that I purchased from a government auction. The unit was purchased without any information about its working state and I have fixing various issues with it. Firstly there was a digital logic problem with the keyboard card that was causing the generator to periodically lock up. When this was fixed the unit came to life but the output signal level was down 5dB and this was traced to a bad step attenuator. I attempted a fix but ended up needing to replace this to get the output steps correct. > > I have now turned my attention to calibrating the reference timebase and unfortunately this seems to be well out of spec. > > The timebase is a HP 10544B 10MHz OCXO. The 10MHz output of the oscillator seems to be ~40Hz low. I have attempted to adjust the unit via the multi-turn capacitor adjustment but this seems to be at the end stop and the frequency is still low. So my questions are: > > 1) I assume this module is faulty as it is so far out of spec? The signal generator has been powered off for multiple months but even after having the unit powered on for 6 hours the frequency has not changed significantly. > > 2) How repairable are these modules? I ended up blowing $200 on the replacement step attenuator so will need to save up for any new parts. I can always use an external 10MHz reference but it is always nice to have equipment complete. > > 3) From searching the list it seems that the oven could be what is faulty. Would a cold over account for this much error? If so I will pull the module and check this. > > 4) Finally is the OCXO just worn out and has aged beyond the point it can be adjusted back into spec? > > > Regards, > > Denis > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com
BG
Bruce Griffiths
Tue, Mar 28, 2023 9:52 PM

Measuring the voltage at the oven monitor pin may be useful in checking if the oven heater voltage is as expected.

Bruce

On 29/03/2023 01:41 NZDT Bob Camp via time-nuts time-nuts@lists.febo.com wrote:

Hi

There’s a range of things that could be wrong. Some of them
are repairable.

If you spend some “quality time” shopping, a replacement can be
had for < $50 delivered.

Bob

On Mar 26, 2023, at 6:56 PM, Denis Dowling via time-nuts time-nuts@lists.febo.com wrote:

Hi All,

This is my first time on the list so apologies if I am going over an old topic.

I am repairing a HP 8662A Synthesised Signal Generator that I purchased from a government auction. The unit was purchased without any information about its working state and I have fixing various issues with it. Firstly there was a digital logic problem with the keyboard card that was causing the generator to periodically lock up. When this was fixed the unit came to life but the output signal level was down 5dB and this was traced to a bad step attenuator. I attempted a fix but ended up needing to replace this to get the output steps correct.

I have now turned my attention to calibrating the reference timebase and unfortunately this seems to be well out of spec.

The timebase is a HP 10544B 10MHz OCXO. The 10MHz output of the oscillator seems to be ~40Hz low. I have attempted to adjust the unit via the multi-turn capacitor adjustment but this seems to be at the end stop and the frequency is still low. So my questions are:

  1. I assume this module is faulty as it is so far out of spec? The signal generator has been powered off for multiple months but even after having the unit powered on for 6 hours the frequency has not changed significantly.

  2. How repairable are these modules? I ended up blowing $200 on the replacement step attenuator so will need to save up for any new parts. I can always use an external 10MHz reference but it is always nice to have equipment complete.

  3. From searching the list it seems that the oven could be what is faulty. Would a cold over account for this much error? If so I will pull the module and check this.

  4. Finally is the OCXO just worn out and has aged beyond the point it can be adjusted back into spec?

Regards,

Denis


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

Measuring the voltage at the oven monitor pin may be useful in checking if the oven heater voltage is as expected. Bruce > On 29/03/2023 01:41 NZDT Bob Camp via time-nuts <time-nuts@lists.febo.com> wrote: > > > Hi > > There’s a range of things that *could* be wrong. Some of them > are repairable. > > If you spend some “quality time” shopping, a replacement can be > had for < $50 delivered. > > Bob > > > On Mar 26, 2023, at 6:56 PM, Denis Dowling via time-nuts <time-nuts@lists.febo.com> wrote: > > > > Hi All, > > > > This is my first time on the list so apologies if I am going over an old topic. > > > > I am repairing a HP 8662A Synthesised Signal Generator that I purchased from a government auction. The unit was purchased without any information about its working state and I have fixing various issues with it. Firstly there was a digital logic problem with the keyboard card that was causing the generator to periodically lock up. When this was fixed the unit came to life but the output signal level was down 5dB and this was traced to a bad step attenuator. I attempted a fix but ended up needing to replace this to get the output steps correct. > > > > I have now turned my attention to calibrating the reference timebase and unfortunately this seems to be well out of spec. > > > > The timebase is a HP 10544B 10MHz OCXO. The 10MHz output of the oscillator seems to be ~40Hz low. I have attempted to adjust the unit via the multi-turn capacitor adjustment but this seems to be at the end stop and the frequency is still low. So my questions are: > > > > 1) I assume this module is faulty as it is so far out of spec? The signal generator has been powered off for multiple months but even after having the unit powered on for 6 hours the frequency has not changed significantly. > > > > 2) How repairable are these modules? I ended up blowing $200 on the replacement step attenuator so will need to save up for any new parts. I can always use an external 10MHz reference but it is always nice to have equipment complete. > > > > 3) From searching the list it seems that the oven could be what is faulty. Would a cold over account for this much error? If so I will pull the module and check this. > > > > 4) Finally is the OCXO just worn out and has aged beyond the point it can be adjusted back into spec? > > > > > > Regards, > > > > Denis > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@lists.febo.com > > To unsubscribe send an email to time-nuts-leave@lists.febo.com > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com