With that offset either the oven set point has changed or there has een a
component value shift in the oscillator circuit.
The easiest way to test is to stick in another 10811 if available.
You could also install one of the original 105 style oscillators if you
have one.
Cheers,
Corby
Ok, so I carefully dismantled the OCXO and I may have found the culprit... The screws holding the two transistors to the metal body were both less than finger tight!
Don't know how that happened, so they were cinched up to 5in-lbs with the torque screwdriver as per the service manual, and I'll give the thing a test once it's all back together.
I'll have a dig around to see if I can find a small thermocouple or similar to see what the oven temperature is at now.
Also, I noticed that a 2.0uF, 20% film capacitor in there is looking like it's had a bit of heat-stroke.
It measures 2.119uF, 0.22ohm ESR, with a dissipation of 0.004 @ 40V, @ 1kHz, so it seems to test ok.
Does this look like it needs to be replaced (picture link below)? I can't find the same part on Digikey/Mouser, but it seems like Keysight have them as 'orderable' status.
Details are as follows: P/N: 0160-4947 Part Description: Capacitor-Fixed 2uF +-20PCT 50V polyester metallized THT
https://i.imgur.com/aZ0KUD8.jpg
Thanks!
Jared
‐‐‐‐‐‐‐ Original Message ‐‐‐‐‐‐‐
On Monday, November 29th, 2021 at 06:30, cdelect@juno.com wrote:
With that offset either the oven set point has changed or there has een a
component value shift in the oscillator circuit.
The easiest way to test is to stick in another 10811 if available.
You could also install one of the original 105 style oscillators if you
have one.
Cheers,
Corby
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
Hi
Quick and dirty way to check the oven is to monitor current. It should start out
near the max rated power and drop back after a few minutes.
Bob
On Nov 30, 2021, at 9:36 AM, Jared Cabot via time-nuts time-nuts@lists.febo.com wrote:
Ok, so I carefully dismantled the OCXO and I may have found the culprit... The screws holding the two transistors to the metal body were both less than finger tight!
Don't know how that happened, so they were cinched up to 5in-lbs with the torque screwdriver as per the service manual, and I'll give the thing a test once it's all back together.
I'll have a dig around to see if I can find a small thermocouple or similar to see what the oven temperature is at now.
Also, I noticed that a 2.0uF, 20% film capacitor in there is looking like it's had a bit of heat-stroke.
It measures 2.119uF, 0.22ohm ESR, with a dissipation of 0.004 @ 40V, @ 1kHz, so it seems to test ok.
Does this look like it needs to be replaced (picture link below)? I can't find the same part on Digikey/Mouser, but it seems like Keysight have them as 'orderable' status.
Details are as follows: P/N: 0160-4947 Part Description: Capacitor-Fixed 2uF +-20PCT 50V polyester metallized THT
https://i.imgur.com/aZ0KUD8.jpg
Thanks!
Jared
‐‐‐‐‐‐‐ Original Message ‐‐‐‐‐‐‐
On Monday, November 29th, 2021 at 06:30, cdelect@juno.com wrote:
With that offset either the oven set point has changed or there has een a
component value shift in the oscillator circuit.
The easiest way to test is to stick in another 10811 if available.
You could also install one of the original 105 style oscillators if you
have one.
Cheers,
Corby
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
That film cap does look a little cooked. Since it is a 2uF 20% 50V part a more common 2.2uF 10% 63V part would likely be an acceptable substitute, and basically still within overall tolerance. Judging from the teflon solder cup used I suspect insulation resistance is a critical parameter. Be sure to clean away any flux remaining after soldering.
Bob L.
Sent: Tuesday, November 30, 2021 at 9:36 AM
From: "Jared Cabot via time-nuts" time-nuts@lists.febo.com
To: "Discussion of precise time and frequency measurement" time-nuts@lists.febo.com
Cc: "Jared Cabot" jaredcabot@protonmail.com
Subject: [time-nuts] Re: HP 5065A, no 2nd harmonic.
...
Also, I noticed that a 2.0uF, 20% film capacitor in there is looking like it's had a bit of heat-stroke.
It measures 2.119uF, 0.22ohm ESR, with a dissipation of 0.004 @ 40V, @ 1kHz, so it seems to test ok.
Does this look like it needs to be replaced (picture link below)? I can't find the same part on Digikey/Mouser, but it seems like Keysight have them as 'orderable' status.
Details are as follows: P/N: 0160-4947 Part Description: Capacitor-Fixed 2uF +-20PCT 50V polyester metallized THT
https://i.imgur.com/aZ0KUD8.jpg
Thanks!
Jared
Well, I found it, the thermal fuse is open circuit.
If I briefly bypass it with a jumper wire, the oven pulls the expected current.
Looks like I need something that will trip at roughly around 115C, rated to 1.0A
(HP Part number 10811-80008)
Hopefully I can find something in Akihabara to make it easier to get my hands on ASAP...
Jared
‐‐‐‐‐‐‐ Original Message ‐‐‐‐‐‐‐
On Wednesday, December 1st, 2021 at 00:41, Robert LaJeunesse lajeunesse@mail.com wrote:
That film cap does look a little cooked. Since it is a 2uF 20% 50V part a more common 2.2uF 10% 63V part would likely be an acceptable substitute, and basically still within overall tolerance. Judging from the teflon solder cup used I suspect insulation resistance is a critical parameter. Be sure to clean away any flux remaining after soldering.
Bob L.
Sent: Tuesday, November 30, 2021 at 9:36 AM
From: "Jared Cabot via time-nuts" time-nuts@lists.febo.com
To: "Discussion of precise time and frequency measurement" time-nuts@lists.febo.com
Cc: "Jared Cabot" jaredcabot@protonmail.com
Subject: [time-nuts] Re: HP 5065A, no 2nd harmonic.
...
Also, I noticed that a 2.0uF, 20% film capacitor in there is looking like it's had a bit of heat-stroke.
It measures 2.119uF, 0.22ohm ESR, with a dissipation of 0.004 @ 40V, @ 1kHz, so it seems to test ok.
Does this look like it needs to be replaced (picture link below)? I can't find the same part on Digikey/Mouser, but it seems like Keysight have them as 'orderable' status.
Details are as follows: P/N: 0160-4947 Part Description: Capacitor-Fixed 2uF +-20PCT 50V polyester metallized THT
Thanks!
Jared
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
Do these thermal fuses fail spontaneously or is there a common cause?
If the "heater" transistor mounting screws were quite loose, those
transistors themselves might have overheated and failed. But, a thermal
fuse should fail by high temperature...
Sounds like progress, though. Cold OCXO means low frequency.
Jared, I sent you a private email with some pictures; did you get that?
Larry
On 11/30/2021 8:03 AM, Jared Cabot via time-nuts wrote:
Well, I found it, the thermal fuse is open circuit.
If I briefly bypass it with a jumper wire, the oven pulls the expected current.
Looks like I need something that will trip at roughly around 115C, rated to 1.0A
(HP Part number 10811-80008)
Hopefully I can find something in Akihabara to make it easier to get my hands on ASAP...
Jared
Best wishes,
Larry McDavid W6FUB
Anaheim, California (SE of Los Angeles, near Disneyland)
I was reading in a past thread on the time-nuts mailing list that the thermal fuses can occasionally spontaneously go open circuit with age, far more often than going open due to a real thermal fault.
(I wonder how many have been discarded over the years due to this issue?)
They appear to be constructed from two spring contacts held closed by wax. As the fuse heats up, the wax melts allowing the contacts to permanently open.
It would appear that age can cause the contacts to open too... A common fix is a wire jumper (The fuse is conveniently socketed, so a resistor leg offcut makes a nice jumper).
I plan to source a replacement thermal fuse, just to keep things a bit safer.
Also, I did get the pictures, thanks for that! I replied directly just now.
Jared.
‐‐‐‐‐‐‐ Original Message ‐‐‐‐‐‐‐
On Wednesday, December 1st, 2021 at 02:05, Larry McDavid lmcdavid@lmceng.com wrote:
Do these thermal fuses fail spontaneously or is there a common cause?
If the "heater" transistor mounting screws were quite loose, those
transistors themselves might have overheated and failed. But, a thermal
fuse should fail by high temperature...
Sounds like progress, though. Cold OCXO means low frequency.
Jared, I sent you a private email with some pictures; did you get that?
Larry
On 11/30/2021 8:03 AM, Jared Cabot via time-nuts wrote:
Well, I found it, the thermal fuse is open circuit.
If I briefly bypass it with a jumper wire, the oven pulls the expected current.
Looks like I need something that will trip at roughly around 115C, rated to 1.0A
(HP Part number 10811-80008)
Hopefully I can find something in Akihabara to make it easier to get my hands on ASAP...
Jared
Best wishes,
Larry McDavid W6FUB
Anaheim, California (SE of Los Angeles, near Disneyland)
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
On 11/30/2021 7:19 AM, Bob kb8tq wrote:
Hi
Quick and dirty way to check the oven is to monitor current. It should start out
near the max rated power and drop back after a few minutes.
And it's worth checking the thermal fuse, which I've had fail in a lot
of them.
Dan
Yes the fuses can just fail like any fuse. But with the oven it pays to
very carefully insure they didn't fail because of temp. and that they did
there job.
Regards
Paul
WB8TSL
On Tue, Nov 30, 2021 at 1:09 PM Dan Rae via time-nuts <
time-nuts@lists.febo.com> wrote:
On 11/30/2021 7:19 AM, Bob kb8tq wrote:
Hi
Quick and dirty way to check the oven is to monitor current. It should
start out
near the max rated power and drop back after a few minutes.
And it's worth checking the thermal fuse, which I've had fail in a lot
of them.
Dan
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
Hi
The thermal fuses die on their own far more often than they die protecting
anything. The fuse is there to provide safety is certain odd situations. The
simple answer these days is to replace it with a piece of wire.
Bob
On Nov 30, 2021, at 12:05 PM, Larry McDavid lmcdavid@lmceng.com wrote:
Do these thermal fuses fail spontaneously or is there a common cause?
If the "heater" transistor mounting screws were quite loose, those transistors themselves might have overheated and failed. But, a thermal fuse should fail by high temperature...
Sounds like progress, though. Cold OCXO means low frequency.
Jared, I sent you a private email with some pictures; did you get that?
Larry
On 11/30/2021 8:03 AM, Jared Cabot via time-nuts wrote:
Well, I found it, the thermal fuse is open circuit.
If I briefly bypass it with a jumper wire, the oven pulls the expected current.
Looks like I need something that will trip at roughly around 115C, rated to 1.0A
(HP Part number 10811-80008)
Hopefully I can find something in Akihabara to make it easier to get my hands on ASAP...
Jared
Best wishes,
Larry McDavid W6FUB
Anaheim, California (SE of Los Angeles, near Disneyland)
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
Did you take a picture showing where the thermal fuse is located? I've never opened a 10811.Sounds like a crude device as you described it. I've used thermal fuses as a safety feature for fast-recharge medical NiCd battery packs but those looked like big diodes.LarrySent via the Samsung Galaxy S20
-------- Original message --------From: Jared Cabot via time-nuts time-nuts@lists.febo.com Date: 11/30/21 9:35 AM (GMT-08:00) To: Discussion of precise time and frequency measurement time-nuts@lists.febo.com Cc: Jared Cabot jaredcabot@protonmail.com Subject: [time-nuts] Re: HP 5065A, no 2nd harmonic. I was reading in a past thread on the time-nuts mailing list that the thermal fuses can occasionally spontaneously go open circuit with age, far more often than going open due to a real thermal fault.(I wonder how many have been discarded over the years due to this issue?)They appear to be constructed from two spring contacts held closed by wax. As the fuse heats up, the wax melts allowing the contacts to permanently open.It would appear that age can cause the contacts to open too... A common fix is a wire jumper (The fuse is conveniently socketed, so a resistor leg offcut makes a nice jumper).I plan to source a replacement thermal fuse, just to keep things a bit safer.Also, I did get the pictures, thanks for that! I replied directly just now.Jared.‐‐‐‐‐‐‐ Original Message ‐‐‐‐‐‐‐On Wednesday, December 1st, 2021 at 02:05, Larry McDavid lmcdavid@lmceng.com wrote:> Do these thermal fuses fail spontaneously or is there a common cause?> > If the "heater" transistor mounting screws were quite loose, those> > transistors themselves might have overheated and failed. But, a thermal> > fuse should fail by high temperature...> > Sounds like progress, though. Cold OCXO means low frequency.> > Jared, I sent you a private email with some pictures; did you get that?> > Larry> > On 11/30/2021 8:03 AM, Jared Cabot via time-nuts wrote:> > > Well, I found it, the thermal fuse is open circuit.> > > > If I briefly bypass it with a jumper wire, the oven pulls the expected current.> > > > Looks like I need something that will trip at roughly around 115C, rated to 1.0A> > > > (HP Part number 10811-80008)> > > > Hopefully I can find something in Akihabara to make it easier to get my hands on ASAP...> > > > Jared> > ...> ---> > Best wishes,> > Larry McDavid W6FUB> > Anaheim, California (SE of Los Angeles, near Disneyland)> > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com> > To unsubscribe, go to and follow the instructions there._______________________________________________time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.comTo unsubscribe, go to and follow the instructions there.
On 1/12/21 5:03 am, Jared Cabot via time-nuts wrote:
Hopefully I can find something in Akihabara
Yeah yeah, don't rub it in! :^)
Hi,
Just a little warning.
Included are some pictures of a 10811 where thermal fuse protection have
not functioned.
I do not know why, if the fuse did not work, or somebody replaced it with a
wire, as this was in a HP 8663A that I found in a recycling bin, and I do
not know the history.
Regards,
Askild
On Tue, Nov 30, 2021 at 8:34 PM Bob kb8tq kb8tq@n1k.org wrote:
Hi
The thermal fuses die on their own far more often than they die protecting
anything. The fuse is there to provide safety is certain odd situations.
The
simple answer these days is to replace it with a piece of wire.
Bob
On Nov 30, 2021, at 12:05 PM, Larry McDavid lmcdavid@lmceng.com wrote:
Do these thermal fuses fail spontaneously or is there a common cause?
If the "heater" transistor mounting screws were quite loose, those
transistors themselves might have overheated and failed. But, a thermal
fuse should fail by high temperature...
Sounds like progress, though. Cold OCXO means low frequency.
Jared, I sent you a private email with some pictures; did you get that?
Larry
On 11/30/2021 8:03 AM, Jared Cabot via time-nuts wrote:
Well, I found it, the thermal fuse is open circuit.
If I briefly bypass it with a jumper wire, the oven pulls the expected
current.
Looks like I need something that will trip at roughly around 115C,
rated to 1.0A
(HP Part number 10811-80008)
Hopefully I can find something in Akihabara to make it easier to get my
hands on ASAP...
Jared
Best wishes,
Larry McDavid W6FUB
Anaheim, California (SE of Los Angeles, near Disneyland)
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe
send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
In general it is a very bad idea to defeat safety provisions that a manufacturer put in their product. They are there for a reason, any manufacturer would omit them, if only to save on cost.
I would think no-one on this list would replace a blown fuse in their house fuse box with a nail. Or?
Wilko
On 1 Dec 2021, at 12:13, Askild megafluffy@gmail.com wrote:
Hi,
Just a little warning.
Included are some pictures of a 10811 where thermal fuse protection have
not functioned.
I do not know why, if the fuse did not work, or somebody replaced it with a
wire, as this was in a HP 8663A that I found in a recycling bin, and I do
not know the history.
Regards,
Askild
On Tue, Nov 30, 2021 at 8:34 PM Bob kb8tq kb8tq@n1k.org wrote:
Hi
The thermal fuses die on their own far more often than they die protecting
anything. The fuse is there to provide safety is certain odd situations.
The
simple answer these days is to replace it with a piece of wire.
Bob
On Nov 30, 2021, at 12:05 PM, Larry McDavid lmcdavid@lmceng.com wrote:
Do these thermal fuses fail spontaneously or is there a common cause?
If the "heater" transistor mounting screws were quite loose, those
transistors themselves might have overheated and failed. But, a thermal
fuse should fail by high temperature...
Sounds like progress, though. Cold OCXO means low frequency.
Jared, I sent you a private email with some pictures; did you get that?
Larry
On 11/30/2021 8:03 AM, Jared Cabot via time-nuts wrote:
Well, I found it, the thermal fuse is open circuit.
If I briefly bypass it with a jumper wire, the oven pulls the expected
current.
Looks like I need something that will trip at roughly around 115C,
rated to 1.0A
(HP Part number 10811-80008)
Hopefully I can find something in Akihabara to make it easier to get my
hands on ASAP...
Jared
Best wishes,
Larry McDavid W6FUB
Anaheim, California (SE of Los Angeles, near Disneyland)
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe
send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
And I will bet the smell is as bad as the larger 0106 oven. As I say its in
a sealed bag and not sure how I will recover it. Need the windows open.
The suggestion of buying a replacement is the best even if you do need to
buy 2-3 to locate a good one.
Regards
Paul
On Wed, Dec 1, 2021 at 6:13 AM Askild megafluffy@gmail.com wrote:
Hi,
Just a little warning.
Included are some pictures of a 10811 where thermal fuse protection have
not functioned.
I do not know why, if the fuse did not work, or somebody replaced it with a
wire, as this was in a HP 8663A that I found in a recycling bin, and I do
not know the history.
Regards,
Askild
On Tue, Nov 30, 2021 at 8:34 PM Bob kb8tq kb8tq@n1k.org wrote:
Hi
The thermal fuses die on their own far more often than they die
protecting
anything. The fuse is there to provide safety is certain odd situations.
The
simple answer these days is to replace it with a piece of wire.
Bob
On Nov 30, 2021, at 12:05 PM, Larry McDavid lmcdavid@lmceng.com
wrote:
Do these thermal fuses fail spontaneously or is there a common cause?
If the "heater" transistor mounting screws were quite loose, those
transistors themselves might have overheated and failed. But, a thermal
fuse should fail by high temperature...
Sounds like progress, though. Cold OCXO means low frequency.
Jared, I sent you a private email with some pictures; did you get that?
Larry
On 11/30/2021 8:03 AM, Jared Cabot via time-nuts wrote:
Well, I found it, the thermal fuse is open circuit.
If I briefly bypass it with a jumper wire, the oven pulls the expected
current.
Looks like I need something that will trip at roughly around 115C,
rated to 1.0A
(HP Part number 10811-80008)
Hopefully I can find something in Akihabara to make it easier to get
my
hands on ASAP...
Jared
Best wishes,
Larry McDavid W6FUB
Anaheim, California (SE of Los Angeles, near Disneyland)
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe
send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe
send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
Hi
The last time I went 10811 shopping, I found a seller with a lot of them.
His price was right ($40 each) if you bought a batch at a time. I tested
them all for ADEV / phase noise and the other basic bench stuff. I did
not run any of them over temperature ( but neither did HP …. ).
Net result was about 5% that had some significant issue ( = you could
not use them in a counter). Maybe another 20% had phase noise or
ADEV that was in spec, but not that great. The balance ranged from pretty
good down a few with high parts in 10^-13 for ADEV.
The real issue is finding a group that has not already been sorted. There
are folks who run through and do the same sort of testing. The worst
1/3 of the group then goes back on eBay …. (including the ones with
significant issues).
Bob
On Dec 1, 2021, at 9:11 AM, paul swed paulswedb@gmail.com wrote:
And I will bet the smell is as bad as the larger 0106 oven. As I say its in
a sealed bag and not sure how I will recover it. Need the windows open.
The suggestion of buying a replacement is the best even if you do need to
buy 2-3 to locate a good one.
Regards
Paul
On Wed, Dec 1, 2021 at 6:13 AM Askild megafluffy@gmail.com wrote:
Hi,
Just a little warning.
Included are some pictures of a 10811 where thermal fuse protection have
not functioned.
I do not know why, if the fuse did not work, or somebody replaced it with a
wire, as this was in a HP 8663A that I found in a recycling bin, and I do
not know the history.
Regards,
Askild
On Tue, Nov 30, 2021 at 8:34 PM Bob kb8tq kb8tq@n1k.org wrote:
Hi
The thermal fuses die on their own far more often than they die
protecting
anything. The fuse is there to provide safety is certain odd situations.
The
simple answer these days is to replace it with a piece of wire.
Bob
On Nov 30, 2021, at 12:05 PM, Larry McDavid lmcdavid@lmceng.com
wrote:
Do these thermal fuses fail spontaneously or is there a common cause?
If the "heater" transistor mounting screws were quite loose, those
transistors themselves might have overheated and failed. But, a thermal
fuse should fail by high temperature...
Sounds like progress, though. Cold OCXO means low frequency.
Jared, I sent you a private email with some pictures; did you get that?
Larry
On 11/30/2021 8:03 AM, Jared Cabot via time-nuts wrote:
Well, I found it, the thermal fuse is open circuit.
If I briefly bypass it with a jumper wire, the oven pulls the expected
current.
Looks like I need something that will trip at roughly around 115C,
rated to 1.0A
(HP Part number 10811-80008)
Hopefully I can find something in Akihabara to make it easier to get
my
hands on ASAP...
Jared
Best wishes,
Larry McDavid W6FUB
Anaheim, California (SE of Los Angeles, near Disneyland)
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe
send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe
send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
Hi
The reason for the thermal fuse (which is not a feature of other OCXO’s)
is the 10811’s deployment in a wide range of devices. There are a few that
do have restrictions on them. The 5065 is not one of those devices.
The protection provided is minimal. The pass transistor will fail in a runaway
situation. Unless you happen to be on a submarine or something similar
you are not creating a significant problem when it does.
The thermal fuse significantly degrades the reliability of the device. It is by
far the most likely part in the entire OCXO to fail. Putting up with this failure
process (and the cost of the fuses) is not reasonable if the benefit is quite
small.
Indeed there was some beer involved in the discussion, but HP OCXO
engineering was onboard with the above information last time we chatted
about it. That would have been in the early 1990’s …. These failures have
been going on for a long time.
Bob
On Dec 1, 2021, at 7:41 AM, Wilko Bulte wkb@xs4all.nl wrote:
In general it is a very bad idea to defeat safety provisions that a manufacturer put in their product. They are there for a reason, any manufacturer would omit them, if only to save on cost.
I would think no-one on this list would replace a blown fuse in their house fuse box with a nail. Or?
Wilko
On 1 Dec 2021, at 12:13, Askild megafluffy@gmail.com wrote:
Hi,
Just a little warning.
Included are some pictures of a 10811 where thermal fuse protection have
not functioned.
I do not know why, if the fuse did not work, or somebody replaced it with a
wire, as this was in a HP 8663A that I found in a recycling bin, and I do
not know the history.
Regards,
Askild
On Tue, Nov 30, 2021 at 8:34 PM Bob kb8tq kb8tq@n1k.org wrote:
Hi
The thermal fuses die on their own far more often than they die protecting
anything. The fuse is there to provide safety is certain odd situations.
The
simple answer these days is to replace it with a piece of wire.
Bob
On Nov 30, 2021, at 12:05 PM, Larry McDavid lmcdavid@lmceng.com wrote:
Do these thermal fuses fail spontaneously or is there a common cause?
If the "heater" transistor mounting screws were quite loose, those
transistors themselves might have overheated and failed. But, a thermal
fuse should fail by high temperature...
Sounds like progress, though. Cold OCXO means low frequency.
Jared, I sent you a private email with some pictures; did you get that?
Larry
On 11/30/2021 8:03 AM, Jared Cabot via time-nuts wrote:
Well, I found it, the thermal fuse is open circuit.
If I briefly bypass it with a jumper wire, the oven pulls the expected
current.
Looks like I need something that will trip at roughly around 115C,
rated to 1.0A
(HP Part number 10811-80008)
Hopefully I can find something in Akihabara to make it easier to get my
hands on ASAP...
Jared
Best wishes,
Larry McDavid W6FUB
Anaheim, California (SE of Los Angeles, near Disneyland)
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe
send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
<P1010309_s.jpg><P1010305_s.jpg><P1010306_s.jpg>
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
Aha! That is very interesting background information! It nicely puts your comment in perspective to "bridge the fuse with a wire and be done with it".
Big grin here when reading the "beer was involved" comment. I have used a similar approach on my business visits to HP CXO (well after T&M was spun off) ;-). A bit like a certain "beer ware license" (hi Poul).
Wilko
On 1 Dec 2021, at 20:20, Bob kb8tq kb8tq@n1k.org wrote:
Hi
The reason for the thermal fuse (which is not a feature of other OCXO’s)
is the 10811’s deployment in a wide range of devices. There are a few that
do have restrictions on them. The 5065 is not one of those devices.
The protection provided is minimal. The pass transistor will fail in a runaway
situation. Unless you happen to be on a submarine or something similar
you are not creating a significant problem when it does.
The thermal fuse significantly degrades the reliability of the device. It is by
far the most likely part in the entire OCXO to fail. Putting up with this failure
process (and the cost of the fuses) is not reasonable if the benefit is quite
small.
Indeed there was some beer involved in the discussion, but HP OCXO
engineering was onboard with the above information last time we chatted
about it. That would have been in the early 1990’s …. These failures have
been going on for a long time.
Bob
On Dec 1, 2021, at 7:41 AM, Wilko Bulte wkb@xs4all.nl wrote:
In general it is a very bad idea to defeat safety provisions that a manufacturer put in their product. They are there for a reason, any manufacturer would omit them, if only to save on cost.
I would think no-one on this list would replace a blown fuse in their house fuse box with a nail. Or?
Wilko
On 1 Dec 2021, at 12:13, Askild megafluffy@gmail.com wrote:
Hi,
Just a little warning.
Included are some pictures of a 10811 where thermal fuse protection have
not functioned.
I do not know why, if the fuse did not work, or somebody replaced it with a
wire, as this was in a HP 8663A that I found in a recycling bin, and I do
not know the history.
Regards,
Askild
On Tue, Nov 30, 2021 at 8:34 PM Bob kb8tq kb8tq@n1k.org wrote:
Hi
The thermal fuses die on their own far more often than they die protecting
anything. The fuse is there to provide safety is certain odd situations.
The
simple answer these days is to replace it with a piece of wire.
Bob
On Nov 30, 2021, at 12:05 PM, Larry McDavid lmcdavid@lmceng.com wrote:
Do these thermal fuses fail spontaneously or is there a common cause?
If the "heater" transistor mounting screws were quite loose, those
transistors themselves might have overheated and failed. But, a thermal
fuse should fail by high temperature...
Sounds like progress, though. Cold OCXO means low frequency.
Jared, I sent you a private email with some pictures; did you get that?
Larry
On 11/30/2021 8:03 AM, Jared Cabot via time-nuts wrote:
Well, I found it, the thermal fuse is open circuit.
If I briefly bypass it with a jumper wire, the oven pulls the expected
current.
Looks like I need something that will trip at roughly around 115C,
rated to 1.0A
(HP Part number 10811-80008)
Hopefully I can find something in Akihabara to make it easier to get my
hands on ASAP...
Jared
Best wishes,
Larry McDavid W6FUB
Anaheim, California (SE of Los Angeles, near Disneyland)
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe
send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
<P1010309_s.jpg><P1010305_s.jpg><P1010306_s.jpg>
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
Well, yay...
I put the OCXO back in the 5065A, and got a good 2nd Harmonic reading. So I started going through the alignment procedures and somehow I now have no 2nd harmonic again, and I can't get it back...
I'm starting to lose interest in this thing, seeing as I have the Leo Bodnar GPS standard anyway that I don't have to sit for hours twiddling dials and poking things just to get it maybe sort of working.
I might give this one more chance before I sell it off to someone with a little more patience, or who prefers fixing gear instead of actually using it. I think I've had this thing working for all of a couple days in the months I've owned it, whereas the GPS standard is the size of a pack of cards and has never failed in the same time.
Jared.
‐‐‐‐‐‐‐ Original Message ‐‐‐‐‐‐‐
On Thursday, December 2nd, 2021 at 04:19, Bob kb8tq kb8tq@n1k.org wrote:
Hi
The reason for the thermal fuse (which is not a feature of other OCXO’s)
is the 10811’s deployment in a wide range of devices. There are a few that
do have restrictions on them. The 5065 is not one of those devices.
The protection provided is minimal. The pass transistor will fail in a runaway
situation. Unless you happen to be on a submarine or something similar
you are not creating a significant problem when it does.
The thermal fuse significantly degrades the reliability of the device. It is by
far the most likely part in the entire OCXO to fail. Putting up with this failure
process (and the cost of the fuses) is not reasonable if the benefit is quite
small.
Indeed there was some beer involved in the discussion, but HP OCXO
engineering was onboard with the above information last time we chatted
about it. That would have been in the early 1990’s …. These failures have
been going on for a long time.
Bob
On Dec 1, 2021, at 7:41 AM, Wilko Bulte wkb@xs4all.nl wrote:
In general it is a very bad idea to defeat safety provisions that a manufacturer put in their product. They are there for a reason, any manufacturer would omit them, if only to save on cost.
I would think no-one on this list would replace a blown fuse in their house fuse box with a nail. Or?
Wilko
On 1 Dec 2021, at 12:13, Askild megafluffy@gmail.com wrote:
Hi,
Just a little warning.
Included are some pictures of a 10811 where thermal fuse protection have
not functioned.
I do not know why, if the fuse did not work, or somebody replaced it with a
wire, as this was in a HP 8663A that I found in a recycling bin, and I do
not know the history.
Regards,
Askild
On Tue, Nov 30, 2021 at 8:34 PM Bob kb8tq kb8tq@n1k.org wrote:
Hi
The thermal fuses die on their own far more often than they die protecting
anything. The fuse is there to provide safety is certain odd situations.
The
simple answer these days is to replace it with a piece of wire.
Bob
On Nov 30, 2021, at 12:05 PM, Larry McDavid lmcdavid@lmceng.com wrote:
Do these thermal fuses fail spontaneously or is there a common cause?
If the "heater" transistor mounting screws were quite loose, those
transistors themselves might have overheated and failed. But, a thermal
fuse should fail by high temperature...
Sounds like progress, though. Cold OCXO means low frequency.
Jared, I sent you a private email with some pictures; did you get that?
Larry
On 11/30/2021 8:03 AM, Jared Cabot via time-nuts wrote:
Well, I found it, the thermal fuse is open circuit.
If I briefly bypass it with a jumper wire, the oven pulls the expected
current.
Looks like I need something that will trip at roughly around 115C,
rated to 1.0A
(HP Part number 10811-80008)
Hopefully I can find something in Akihabara to make it easier to get my
hands on ASAP...
Jared
...
--
Best wishes,
Larry McDavid W6FUB
Anaheim, California (SE of Los Angeles, near Disneyland)
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe
send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
<P1010309_s.jpg><P1010305_s.jpg><P1010306_s.jpg>
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.
Don't give up now, Jared! You must be so close to getting that puppy running correctly.
On 2/12/21 9:16 pm, Jared Cabot via time-nuts wrote:
Well, yay...
I put the OCXO back in the 5065A, and got a good 2nd Harmonic reading. So I started going through the alignment procedures and somehow I now have no 2nd harmonic again, and I can't get it back...
I'm starting to lose interest in this thing, seeing as I have the Leo Bodnar GPS standard anyway that I don't have to sit for hours twiddling dials and poking things just to get it maybe sort of working.
I might give this one more chance before I sell it off to someone with a little more patience, or who prefers fixing gear instead of actually using it. I think I've had this thing working for all of a couple days in the months I've owned it, whereas the GPS standard is the size of a pack of cards and has never failed in the same time.