Fluke 732A Questions

JL
J. L. Trantham
Sat, Dec 28, 2013 10:51 PM

I have reviewed the prior postings on Volt-nuts (a very rewarding review,
BTW)) regarding the 732A and have two questions:

  1.    Has anyone determined the part number, or a source, for the mating
    

plug to J10, the external power connector for the 732A-7005 battery pack?  I
noted some postings recently about this but did not find a definitive
response regarding the identity of this plug.  The alternative would appear
to be a complete replacement of J10 and its plug, as long is it all fits in
the opening in the panel.  I hate to 'bore holes' in vintage equipment.

  1.   The oven thermistor in my 'new to me' unit measures 4229 ohms (+/-
    

an ohm or two) after the unit has warmed up for a week or so.  I note the
manual refers to 3K to 4K for the value of this thermistor when the unit is
'stable' as well as other's posting values in the mid 3500's ohms for their
units.  Should I be concerned?  The unit seems to be stable to within about
2 uV over about a week (as measured by my 3458A - see below).  Should I open
the unit and try to measure the oven temperature or just be satisfied that
the unit seems to be working?

I had to replace the four 6V 4AH SLA batteries and they charged up
appropriately as judged by the front panel LED's.  I had to remove the
'jumper' for the '40' option on the A7 board and connect the jumper to the
'20' and '10' options (total of '30') in order to get the unit to adjust to
10.0000000 VDC on my 'Agilent In Cal'd' 3458A.

The need to change the jumpers, perhaps, could be just an ageing issue or,
on the other hand, a 'temperature' issue with the oven.

Should I open the unit and directly measure the temperature (supposedly
about 48 degrees C) or just be satisfied with what I have?

My recently added 735C also needed moving some 'jumpers' in order to get it
'on scale', as determined by my 3458A, although it's thermistor measures
about 3330 ohms (after being on for several weeks).

Thanks for everyone's help.

Joe

I have reviewed the prior postings on Volt-nuts (a very rewarding review, BTW)) regarding the 732A and have two questions: 1. Has anyone determined the part number, or a source, for the mating plug to J10, the external power connector for the 732A-7005 battery pack? I noted some postings recently about this but did not find a definitive response regarding the identity of this plug. The alternative would appear to be a complete replacement of J10 and its plug, as long is it all fits in the opening in the panel. I hate to 'bore holes' in vintage equipment. 2. The oven thermistor in my 'new to me' unit measures 4229 ohms (+/- an ohm or two) after the unit has warmed up for a week or so. I note the manual refers to 3K to 4K for the value of this thermistor when the unit is 'stable' as well as other's posting values in the mid 3500's ohms for their units. Should I be concerned? The unit seems to be stable to within about 2 uV over about a week (as measured by my 3458A - see below). Should I open the unit and try to measure the oven temperature or just be satisfied that the unit seems to be working? I had to replace the four 6V 4AH SLA batteries and they charged up appropriately as judged by the front panel LED's. I had to remove the 'jumper' for the '40' option on the A7 board and connect the jumper to the '20' and '10' options (total of '30') in order to get the unit to adjust to 10.0000000 VDC on my 'Agilent In Cal'd' 3458A. The need to change the jumpers, perhaps, could be just an ageing issue or, on the other hand, a 'temperature' issue with the oven. Should I open the unit and directly measure the temperature (supposedly about 48 degrees C) or just be satisfied with what I have? My recently added 735C also needed moving some 'jumpers' in order to get it 'on scale', as determined by my 3458A, although it's thermistor measures about 3330 ohms (after being on for several weeks). Thanks for everyone's help. Joe
R
R.Phillips
Sun, Dec 29, 2013 2:39 PM

Hi Joe
I also obtained a Fluke 732A back in October last, and I have fitted a new
set of SLA batteries and it is working just fine - I am still very impressed
by its accuracy and stability, and its build quality.
First, the external 3-pin power socket is a HYPERTRONICS 'D' type connector
(female) type 100-166 - but I have not found one to date !
I also expected the temperature monitoring thermistor to give a 3 to 4 K
ohms, as indicated in the manual P/N 645051 dated May 1983 - but I guess
that you have a later model as I do, mine being serial # 4845***, some
modifications were made,and I understand that this included a change of
thermistor, so the standard reading is in keeping with your result. My unit
gives a very stable 4.520 K ohms, being between 4.52080 and 4.52090 K ohms.
Judging by your findings, you are also using the 1983 copy of the User
Manual - we could both use a more up to date issue.
I am awaiting the repair of my 3458A - so I am having to rely upon a
recently calibrated Keithley 2015 and my good old 3456A - but they both give
very impressive results - The Keithley giving 1.000007 volts for the 1 volt
output, and 10.00000 volts for the 10 volt output (how I miss that final
digit).
Looking forward to your further results and any other owners comments.
Best regards

Roy

-----Original Message-----
From: J. L. Trantham
Sent: Saturday, December 28, 2013 10:51 PM
To: volt-nuts@febo.com
Subject: [volt-nuts] Fluke 732A Questions

I have reviewed the prior postings on Volt-nuts (a very rewarding review,
BTW)) regarding the 732A and have two questions:

  1.    Has anyone determined the part number, or a source, for the mating
    

plug to J10, the external power connector for the 732A-7005 battery pack?  I
noted some postings recently about this but did not find a definitive
response regarding the identity of this plug.  The alternative would appear
to be a complete replacement of J10 and its plug, as long is it all fits in
the opening in the panel.  I hate to 'bore holes' in vintage equipment.

  1.   The oven thermistor in my 'new to me' unit measures 4229 ohms (+/-
    

an ohm or two) after the unit has warmed up for a week or so.  I note the
manual refers to 3K to 4K for the value of this thermistor when the unit is
'stable' as well as other's posting values in the mid 3500's ohms for their
units.  Should I be concerned?  The unit seems to be stable to within about
2 uV over about a week (as measured by my 3458A - see below).  Should I open
the unit and try to measure the oven temperature or just be satisfied that
the unit seems to be working?

I had to replace the four 6V 4AH SLA batteries and they charged up
appropriately as judged by the front panel LED's.  I had to remove the
'jumper' for the '40' option on the A7 board and connect the jumper to the
'20' and '10' options (total of '30') in order to get the unit to adjust to
10.0000000 VDC on my 'Agilent In Cal'd' 3458A.

The need to change the jumpers, perhaps, could be just an ageing issue or,
on the other hand, a 'temperature' issue with the oven.

Should I open the unit and directly measure the temperature (supposedly
about 48 degrees C) or just be satisfied with what I have?

My recently added 735C also needed moving some 'jumpers' in order to get it
'on scale', as determined by my 3458A, although it's thermistor measures
about 3330 ohms (after being on for several weeks).

Thanks for everyone's help.

Joe


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi Joe I also obtained a Fluke 732A back in October last, and I have fitted a new set of SLA batteries and it is working just fine - I am still very impressed by its accuracy and stability, and its build quality. First, the external 3-pin power socket is a HYPERTRONICS 'D' type connector (female) type 100-166 - but I have not found one to date ! I also expected the temperature monitoring thermistor to give a 3 to 4 K ohms, as indicated in the manual P/N 645051 dated May 1983 - but I guess that you have a later model as I do, mine being serial # 4845***, some modifications were made,and I understand that this included a change of thermistor, so the standard reading is in keeping with your result. My unit gives a very stable 4.520 K ohms, being between 4.52080 and 4.52090 K ohms. Judging by your findings, you are also using the 1983 copy of the User Manual - we could both use a more up to date issue. I am awaiting the repair of my 3458A - so I am having to rely upon a recently calibrated Keithley 2015 and my good old 3456A - but they both give very impressive results - The Keithley giving 1.000007 volts for the 1 volt output, and 10.00000 volts for the 10 volt output (how I miss that final digit). Looking forward to your further results and any other owners comments. Best regards Roy -----Original Message----- From: J. L. Trantham Sent: Saturday, December 28, 2013 10:51 PM To: volt-nuts@febo.com Subject: [volt-nuts] Fluke 732A Questions I have reviewed the prior postings on Volt-nuts (a very rewarding review, BTW)) regarding the 732A and have two questions: 1. Has anyone determined the part number, or a source, for the mating plug to J10, the external power connector for the 732A-7005 battery pack? I noted some postings recently about this but did not find a definitive response regarding the identity of this plug. The alternative would appear to be a complete replacement of J10 and its plug, as long is it all fits in the opening in the panel. I hate to 'bore holes' in vintage equipment. 2. The oven thermistor in my 'new to me' unit measures 4229 ohms (+/- an ohm or two) after the unit has warmed up for a week or so. I note the manual refers to 3K to 4K for the value of this thermistor when the unit is 'stable' as well as other's posting values in the mid 3500's ohms for their units. Should I be concerned? The unit seems to be stable to within about 2 uV over about a week (as measured by my 3458A - see below). Should I open the unit and try to measure the oven temperature or just be satisfied that the unit seems to be working? I had to replace the four 6V 4AH SLA batteries and they charged up appropriately as judged by the front panel LED's. I had to remove the 'jumper' for the '40' option on the A7 board and connect the jumper to the '20' and '10' options (total of '30') in order to get the unit to adjust to 10.0000000 VDC on my 'Agilent In Cal'd' 3458A. The need to change the jumpers, perhaps, could be just an ageing issue or, on the other hand, a 'temperature' issue with the oven. Should I open the unit and directly measure the temperature (supposedly about 48 degrees C) or just be satisfied with what I have? My recently added 735C also needed moving some 'jumpers' in order to get it 'on scale', as determined by my 3458A, although it's thermistor measures about 3330 ohms (after being on for several weeks). Thanks for everyone's help. Joe _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
JL
J. L. Trantham
Sun, Dec 29, 2013 6:12 PM

Roy,

Thanks for the info.

Looking at the Hypertronics catalog, it would appear that the panel mounted
socket (with male, solder cup, pins) is:

P/N D01EEB306MSUTH

And the mating plug would be:

P/N D01PB306FSUTAH

However, the dimensions are a bit off and the appearance of the panel
mounted socket is not the same as what is on my 732A, J10, on the back of
the 732A-7005 battery module.  My connector is exactly as shown in the
change sheets with a dimension of 0.475 inches side to side and 0.500 inches
top to bottom (as viewed from the back of the unit) rather than 0.512 x
0.512 as suggested in the catalog.

However, the diameter of the mounting hole and the presence of a 'notch'
appears to be as shown in the catalog, thus suggesting that it, indeed, is a
series D01 part.  The diameter of the threaded part of J10 is 0.410 inches
and the catalog calls for a hole diameter of 0.441 inches.

One other observation is the presence of a 'symbol' on the bottom edge of
the socket (as viewed from below) that looks like a stylized 'FRE',  The 'F'
perhaps could be an 'A' and the 'E' perhaps could be a 'B'.  The symbol
starts low and ends low with a 'peak' of the symbol in the middle of the
'R', sort of a 'roof' shape to the top of the symbol and flat on the bottom
of the symbol.

Not sure what that means but I have a picture of it if anyone thinks it
would help.  It does not look like the stylized 'H' symbol shown in the
Hpertronic catalog.  Perhaps the company was something else before they
became Hypertronics.

I looked through the change sheets for the 732A and could find no mention of
the value of the thermistor resistance at temperature, only comments about
what its stability should be.  Knowing that you have a value in the mid 4K
range is reassuring.  My S/N starts with 3.

Right now, it indicates 10.0000005 VDC on the 3458A and 4230 ohms on the
Fluke 8050A.

Thanks for your help.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of R.Phillips
Sent: Sunday, December 29, 2013 8:40 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions

Hi Joe
I also obtained a Fluke 732A back in October last, and I have fitted a new
set of SLA batteries and it is working just fine - I am still very impressed
by its accuracy and stability, and its build quality.
First, the external 3-pin power socket is a HYPERTRONICS 'D' type connector
(female) type 100-166 - but I have not found one to date !
I also expected the temperature monitoring thermistor to give a 3 to 4 K
ohms, as indicated in the manual P/N 645051 dated May 1983 - but I guess
that you have a later model as I do, mine being serial # 4845***, some
modifications were made,and I understand that this included a change of
thermistor, so the standard reading is in keeping with your result. My unit
gives a very stable 4.520 K ohms, being between 4.52080 and 4.52090 K ohms.
Judging by your findings, you are also using the 1983 copy of the User
Manual - we could both use a more up to date issue.
I am awaiting the repair of my 3458A - so I am having to rely upon a
recently calibrated Keithley 2015 and my good old 3456A - but they both give
very impressive results - The Keithley giving 1.000007 volts for the 1 volt
output, and 10.00000 volts for the 10 volt output (how I miss that final
digit).
Looking forward to your further results and any other owners comments.
Best regards

Roy

-----Original Message-----
From: J. L. Trantham
Sent: Saturday, December 28, 2013 10:51 PM
To: volt-nuts@febo.com
Subject: [volt-nuts] Fluke 732A Questions

I have reviewed the prior postings on Volt-nuts (a very rewarding review,
BTW)) regarding the 732A and have two questions:

  1.    Has anyone determined the part number, or a source, for the mating
    

plug to J10, the external power connector for the 732A-7005 battery pack?  I
noted some postings recently about this but did not find a definitive
response regarding the identity of this plug.  The alternative would appear
to be a complete replacement of J10 and its plug, as long is it all fits in
the opening in the panel.  I hate to 'bore holes' in vintage equipment.

  1.   The oven thermistor in my 'new to me' unit measures 4229 ohms (+/-
    

an ohm or two) after the unit has warmed up for a week or so.  I note the
manual refers to 3K to 4K for the value of this thermistor when the unit is
'stable' as well as other's posting values in the mid 3500's ohms for their
units.  Should I be concerned?  The unit seems to be stable to within about
2 uV over about a week (as measured by my 3458A - see below).  Should I open
the unit and try to measure the oven temperature or just be satisfied that
the unit seems to be working?

I had to replace the four 6V 4AH SLA batteries and they charged up
appropriately as judged by the front panel LED's.  I had to remove the
'jumper' for the '40' option on the A7 board and connect the jumper to the
'20' and '10' options (total of '30') in order to get the unit to adjust to
10.0000000 VDC on my 'Agilent In Cal'd' 3458A.

The need to change the jumpers, perhaps, could be just an ageing issue or,
on the other hand, a 'temperature' issue with the oven.

Should I open the unit and directly measure the temperature (supposedly
about 48 degrees C) or just be satisfied with what I have?

My recently added 735C also needed moving some 'jumpers' in order to get it
'on scale', as determined by my 3458A, although it's thermistor measures
about 3330 ohms (after being on for several weeks).

Thanks for everyone's help.

Joe


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Roy, Thanks for the info. Looking at the Hypertronics catalog, it would appear that the panel mounted socket (with male, solder cup, pins) is: P/N D01EEB306MSUTH And the mating plug would be: P/N D01PB306FSUTAH However, the dimensions are a bit off and the appearance of the panel mounted socket is not the same as what is on my 732A, J10, on the back of the 732A-7005 battery module. My connector is exactly as shown in the change sheets with a dimension of 0.475 inches side to side and 0.500 inches top to bottom (as viewed from the back of the unit) rather than 0.512 x 0.512 as suggested in the catalog. However, the diameter of the mounting hole and the presence of a 'notch' appears to be as shown in the catalog, thus suggesting that it, indeed, is a series D01 part. The diameter of the threaded part of J10 is 0.410 inches and the catalog calls for a hole diameter of 0.441 inches. One other observation is the presence of a 'symbol' on the bottom edge of the socket (as viewed from below) that looks like a stylized 'FRE', The 'F' perhaps could be an 'A' and the 'E' perhaps could be a 'B'. The symbol starts low and ends low with a 'peak' of the symbol in the middle of the 'R', sort of a 'roof' shape to the top of the symbol and flat on the bottom of the symbol. Not sure what that means but I have a picture of it if anyone thinks it would help. It does not look like the stylized 'H' symbol shown in the Hpertronic catalog. Perhaps the company was something else before they became Hypertronics. I looked through the change sheets for the 732A and could find no mention of the value of the thermistor resistance at temperature, only comments about what its stability should be. Knowing that you have a value in the mid 4K range is reassuring. My S/N starts with 3. Right now, it indicates 10.0000005 VDC on the 3458A and 4230 ohms on the Fluke 8050A. Thanks for your help. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of R.Phillips Sent: Sunday, December 29, 2013 8:40 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 732A Questions Hi Joe I also obtained a Fluke 732A back in October last, and I have fitted a new set of SLA batteries and it is working just fine - I am still very impressed by its accuracy and stability, and its build quality. First, the external 3-pin power socket is a HYPERTRONICS 'D' type connector (female) type 100-166 - but I have not found one to date ! I also expected the temperature monitoring thermistor to give a 3 to 4 K ohms, as indicated in the manual P/N 645051 dated May 1983 - but I guess that you have a later model as I do, mine being serial # 4845***, some modifications were made,and I understand that this included a change of thermistor, so the standard reading is in keeping with your result. My unit gives a very stable 4.520 K ohms, being between 4.52080 and 4.52090 K ohms. Judging by your findings, you are also using the 1983 copy of the User Manual - we could both use a more up to date issue. I am awaiting the repair of my 3458A - so I am having to rely upon a recently calibrated Keithley 2015 and my good old 3456A - but they both give very impressive results - The Keithley giving 1.000007 volts for the 1 volt output, and 10.00000 volts for the 10 volt output (how I miss that final digit). Looking forward to your further results and any other owners comments. Best regards Roy -----Original Message----- From: J. L. Trantham Sent: Saturday, December 28, 2013 10:51 PM To: volt-nuts@febo.com Subject: [volt-nuts] Fluke 732A Questions I have reviewed the prior postings on Volt-nuts (a very rewarding review, BTW)) regarding the 732A and have two questions: 1. Has anyone determined the part number, or a source, for the mating plug to J10, the external power connector for the 732A-7005 battery pack? I noted some postings recently about this but did not find a definitive response regarding the identity of this plug. The alternative would appear to be a complete replacement of J10 and its plug, as long is it all fits in the opening in the panel. I hate to 'bore holes' in vintage equipment. 2. The oven thermistor in my 'new to me' unit measures 4229 ohms (+/- an ohm or two) after the unit has warmed up for a week or so. I note the manual refers to 3K to 4K for the value of this thermistor when the unit is 'stable' as well as other's posting values in the mid 3500's ohms for their units. Should I be concerned? The unit seems to be stable to within about 2 uV over about a week (as measured by my 3458A - see below). Should I open the unit and try to measure the oven temperature or just be satisfied that the unit seems to be working? I had to replace the four 6V 4AH SLA batteries and they charged up appropriately as judged by the front panel LED's. I had to remove the 'jumper' for the '40' option on the A7 board and connect the jumper to the '20' and '10' options (total of '30') in order to get the unit to adjust to 10.0000000 VDC on my 'Agilent In Cal'd' 3458A. The need to change the jumpers, perhaps, could be just an ageing issue or, on the other hand, a 'temperature' issue with the oven. Should I open the unit and directly measure the temperature (supposedly about 48 degrees C) or just be satisfied with what I have? My recently added 735C also needed moving some 'jumpers' in order to get it 'on scale', as determined by my 3458A, although it's thermistor measures about 3330 ohms (after being on for several weeks). Thanks for everyone's help. Joe _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
RA
Robert Atkinson
Sun, Dec 29, 2013 7:14 PM

Hi Joe,
As Roy has already said, this is a hypertronics connector. I've had one other piece of equipment with this type of connector, a nitrogen laser. See < http://www.thinksrs.com/products/NL100.htm > The downloadable user manual has details. Sorry, I don't have any, or a cheap source.

Robert G8RPI.


From: J. L. Trantham jltran@att.net
To: volt-nuts@febo.com
Sent: Saturday, 28 December 2013, 22:51
Subject: [volt-nuts] Fluke 732A Questions

I have reviewed the prior postings on Volt-nuts (a very rewarding review,
BTW)) regarding the 732A and have two questions:

1.        Has anyone determined the part number, or a source, for the mating
plug to J10, the external power connector for the 732A-7005 battery pack?  I
noted some postings recently about this but did not find a definitive
response regarding the identity of this plug.  The alternative would appear
to be a complete replacement of J10 and its plug, as long is it all fits in
the opening in the panel.  I hate to 'bore holes' in vintage equipment.

2.       The oven thermistor in my 'new to me' unit measures 4229 ohms (+/-
an ohm or two) after the unit has warmed up for a week or so.  I note the
manual refers to 3K to 4K for the value of this thermistor when the unit is
'stable' as well as other's posting values in the mid 3500's ohms for their
units.  Should I be concerned?  The unit seems to be stable to within about
2 uV over about a week (as measured by my 3458A - see below).  Should I open
the unit and try to measure the oven temperature or just be satisfied that
the unit seems to be working?

I had to replace the four 6V 4AH SLA batteries and they charged up
appropriately as judged by the front panel LED's.  I had to remove the
'jumper' for the '40' option on the A7 board and connect the jumper to the
'20' and '10' options (total of '30') in order to get the unit to adjust to
10.0000000 VDC on my 'Agilent In Cal'd' 3458A. 

The need to change the jumpers, perhaps, could be just an ageing issue or,
on the other hand, a 'temperature' issue with the oven.

Should I open the unit and directly measure the temperature (supposedly
about 48 degrees C) or just be satisfied with what I have?

My recently added 735C also needed moving some 'jumpers' in order to get it
'on scale', as determined by my 3458A, although it's thermistor measures
about 3330 ohms (after being on for several weeks).

Thanks for everyone's help.

Joe


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi Joe, As Roy has already said, this is a hypertronics connector. I've had one other piece of equipment with this type of connector, a nitrogen laser. See < http://www.thinksrs.com/products/NL100.htm > The downloadable user manual has details. Sorry, I don't have any, or a cheap source. Robert G8RPI. ________________________________ From: J. L. Trantham <jltran@att.net> To: volt-nuts@febo.com Sent: Saturday, 28 December 2013, 22:51 Subject: [volt-nuts] Fluke 732A Questions I have reviewed the prior postings on Volt-nuts (a very rewarding review, BTW)) regarding the 732A and have two questions: 1.        Has anyone determined the part number, or a source, for the mating plug to J10, the external power connector for the 732A-7005 battery pack?  I noted some postings recently about this but did not find a definitive response regarding the identity of this plug.  The alternative would appear to be a complete replacement of J10 and its plug, as long is it all fits in the opening in the panel.  I hate to 'bore holes' in vintage equipment. 2.      The oven thermistor in my 'new to me' unit measures 4229 ohms (+/- an ohm or two) after the unit has warmed up for a week or so.  I note the manual refers to 3K to 4K for the value of this thermistor when the unit is 'stable' as well as other's posting values in the mid 3500's ohms for their units.  Should I be concerned?  The unit seems to be stable to within about 2 uV over about a week (as measured by my 3458A - see below).  Should I open the unit and try to measure the oven temperature or just be satisfied that the unit seems to be working? I had to replace the four 6V 4AH SLA batteries and they charged up appropriately as judged by the front panel LED's.  I had to remove the 'jumper' for the '40' option on the A7 board and connect the jumper to the '20' and '10' options (total of '30') in order to get the unit to adjust to 10.0000000 VDC on my 'Agilent In Cal'd' 3458A.  The need to change the jumpers, perhaps, could be just an ageing issue or, on the other hand, a 'temperature' issue with the oven. Should I open the unit and directly measure the temperature (supposedly about 48 degrees C) or just be satisfied with what I have? My recently added 735C also needed moving some 'jumpers' in order to get it 'on scale', as determined by my 3458A, although it's thermistor measures about 3330 ohms (after being on for several weeks). Thanks for everyone's help. Joe _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
JL
J. L. Trantham
Sun, Dec 29, 2013 11:55 PM

Robert,

Thanks for the info.

The picture of the socket on the Nitrogen LASER certainly looks like what
the Hypertronics catalog shows and the part number for the plug is certainly
similar (except the LASER has female connectors and the plug has male
connectors) to what I would think would be the plug for the 732A.  However
the picture of the D01 socket on the LASER is not the same as the socket on
the 732A.  The plug, though, might fit both.

I emailed Hypertronics asking if there is a link between Hypertronics and
'FRE' (whatever that is) and if the part number I came up with is available
or not.  The 'automated' responses mentioned 'out of the office' until Jan 6
or so.

We'll see what response I get.

Did you, perhaps, get a plug with the LASER?  If so, did you see if it would
fit the 732A?

Thanks again for the info.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Robert Atkinson
Sent: Sunday, December 29, 2013 1:14 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions

Hi Joe,
As Roy has already said, this is a hypertronics connector. I've had one
other piece of equipment with this type of connector, a nitrogen laser. See
< http://www.thinksrs.com/products/NL100.htm > The downloadable user manual
has details. Sorry, I don't have any, or a cheap source.

Robert G8RPI.


From: J. L. Trantham jltran@att.net
To: volt-nuts@febo.com
Sent: Saturday, 28 December 2013, 22:51
Subject: [volt-nuts] Fluke 732A Questions

I have reviewed the prior postings on Volt-nuts (a very rewarding review,
BTW)) regarding the 732A and have two questions:

1.        Has anyone determined the part number, or a source, for the mating
plug to J10, the external power connector for the 732A-7005 battery pack?  I
noted some postings recently about this but did not find a definitive
response regarding the identity of this plug.  The alternative would appear
to be a complete replacement of J10 and its plug, as long is it all fits in
the opening in the panel.  I hate to 'bore holes' in vintage equipment.

2.       The oven thermistor in my 'new to me' unit measures 4229 ohms (+/-
an ohm or two) after the unit has warmed up for a week or so.  I note the
manual refers to 3K to 4K for the value of this thermistor when the unit is
'stable' as well as other's posting values in the mid 3500's ohms for their
units.  Should I be concerned?  The unit seems to be stable to within about
2 uV over about a week (as measured by my 3458A - see below).  Should I open
the unit and try to measure the oven temperature or just be satisfied that
the unit seems to be working?

I had to replace the four 6V 4AH SLA batteries and they charged up
appropriately as judged by the front panel LED's.  I had to remove the
'jumper' for the '40' option on the A7 board and connect the jumper to the
'20' and '10' options (total of '30') in order to get the unit to adjust to
10.0000000 VDC on my 'Agilent In Cal'd' 3458A. 

The need to change the jumpers, perhaps, could be just an ageing issue or,
on the other hand, a 'temperature' issue with the oven.

Should I open the unit and directly measure the temperature (supposedly
about 48 degrees C) or just be satisfied with what I have?

My recently added 735C also needed moving some 'jumpers' in order to get it
'on scale', as determined by my 3458A, although it's thermistor measures
about 3330 ohms (after being on for several weeks).

Thanks for everyone's help.

Joe


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Robert, Thanks for the info. The picture of the socket on the Nitrogen LASER certainly looks like what the Hypertronics catalog shows and the part number for the plug is certainly similar (except the LASER has female connectors and the plug has male connectors) to what I would think would be the plug for the 732A. However the picture of the D01 socket on the LASER is not the same as the socket on the 732A. The plug, though, might fit both. I emailed Hypertronics asking if there is a link between Hypertronics and 'FRE' (whatever that is) and if the part number I came up with is available or not. The 'automated' responses mentioned 'out of the office' until Jan 6 or so. We'll see what response I get. Did you, perhaps, get a plug with the LASER? If so, did you see if it would fit the 732A? Thanks again for the info. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Robert Atkinson Sent: Sunday, December 29, 2013 1:14 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 732A Questions Hi Joe, As Roy has already said, this is a hypertronics connector. I've had one other piece of equipment with this type of connector, a nitrogen laser. See < http://www.thinksrs.com/products/NL100.htm > The downloadable user manual has details. Sorry, I don't have any, or a cheap source. Robert G8RPI. ________________________________ From: J. L. Trantham <jltran@att.net> To: volt-nuts@febo.com Sent: Saturday, 28 December 2013, 22:51 Subject: [volt-nuts] Fluke 732A Questions I have reviewed the prior postings on Volt-nuts (a very rewarding review, BTW)) regarding the 732A and have two questions: 1.        Has anyone determined the part number, or a source, for the mating plug to J10, the external power connector for the 732A-7005 battery pack?  I noted some postings recently about this but did not find a definitive response regarding the identity of this plug.  The alternative would appear to be a complete replacement of J10 and its plug, as long is it all fits in the opening in the panel.  I hate to 'bore holes' in vintage equipment. 2.      The oven thermistor in my 'new to me' unit measures 4229 ohms (+/- an ohm or two) after the unit has warmed up for a week or so.  I note the manual refers to 3K to 4K for the value of this thermistor when the unit is 'stable' as well as other's posting values in the mid 3500's ohms for their units.  Should I be concerned?  The unit seems to be stable to within about 2 uV over about a week (as measured by my 3458A - see below).  Should I open the unit and try to measure the oven temperature or just be satisfied that the unit seems to be working? I had to replace the four 6V 4AH SLA batteries and they charged up appropriately as judged by the front panel LED's.  I had to remove the 'jumper' for the '40' option on the A7 board and connect the jumper to the '20' and '10' options (total of '30') in order to get the unit to adjust to 10.0000000 VDC on my 'Agilent In Cal'd' 3458A.  The need to change the jumpers, perhaps, could be just an ageing issue or, on the other hand, a 'temperature' issue with the oven. Should I open the unit and directly measure the temperature (supposedly about 48 degrees C) or just be satisfied with what I have? My recently added 735C also needed moving some 'jumpers' in order to get it 'on scale', as determined by my 3458A, although it's thermistor measures about 3330 ohms (after being on for several weeks). Thanks for everyone's help. Joe _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
TM
T. Micallef
Mon, Dec 30, 2013 12:04 AM

Joe,

I have the external battery pack (732A-7003) that I picked up a while ago.
It is indeed the D01 connector. It is printed clearly on the outside edge
along with the three letters that look like ARB.

The company's website states they are now owned by Smiths Connectors and the
connector is part of Hypertac. No list of distributors was found on the website.

I found a seller Astrex Electronics, but they do not have a price listed for
the female version, just the male. I searched for D01P306FSTAH. This is the
female plug with solder cups.

To date, I have reconditioned three 732A's and they have been operating fine
with little issues, except when I lost power for an extended time and two
batteries in one of the packs died. All four were replaced and they have
been working fine.

The two older 732A's have thermistor values in the 3k-4k range. The newer
unit (based on the newer heater rev) is around 4.25K.

All three had their potentiometers changed on the regulator and
pre-regulator boards. The datecodes on all appeared to be from 1985. The
parts list must have been updated as some of the pots did not have the same
value as the ones installed. I cannot remember which ones were different. I
did not see a mention in the change/errata.

Also, I had to remove the reference ovens to repair broken 10V
potentiometers, as someone had broken one (free spinning) and another unit
appeared to have a noisy wiper so it was replaced as well. The 1V and 1.018V
pots were not changed. The third 10V pot was replaced because I had success
with the other two and it was much easier the third time around. The jumpers
on the all the A7 boards were adjusted.

I highly recommend checking the inline fuse to the heater if you suspect it
is not working. You can find it tucked inside behind the oven. It should be
covered in heat shrink.

Lastly, there was a rev done that does not appear in the downloadable users
manual on Didier's website.  It appears the heaters were separated into two
groups of two instead of one group of four. Some of the parts were mounted
under the oven assembly on the A8 Preheater PCA. If you look at the A4 board
from the top, you will see either one or two transistors mounted to the
frame (Q4,Q14). If you see only one (Q4), chances are you have the newer rev.

I bought the manual recently and just noticed these revs were in my book.
I will try and scan the extra info and upload them.

Todd

Joe, I have the external battery pack (732A-7003) that I picked up a while ago. It is indeed the D01 connector. It is printed clearly on the outside edge along with the three letters that look like ARB. The company's website states they are now owned by Smiths Connectors and the connector is part of Hypertac. No list of distributors was found on the website. I found a seller Astrex Electronics, but they do not have a price listed for the female version, just the male. I searched for D01P306FSTAH. This is the female plug with solder cups. To date, I have reconditioned three 732A's and they have been operating fine with little issues, except when I lost power for an extended time and two batteries in one of the packs died. All four were replaced and they have been working fine. The two older 732A's have thermistor values in the 3k-4k range. The newer unit (based on the newer heater rev) is around 4.25K. All three had their potentiometers changed on the regulator and pre-regulator boards. The datecodes on all appeared to be from 1985. The parts list must have been updated as some of the pots did not have the same value as the ones installed. I cannot remember which ones were different. I did not see a mention in the change/errata. Also, I had to remove the reference ovens to repair broken 10V potentiometers, as someone had broken one (free spinning) and another unit appeared to have a noisy wiper so it was replaced as well. The 1V and 1.018V pots were not changed. The third 10V pot was replaced because I had success with the other two and it was much easier the third time around. The jumpers on the all the A7 boards were adjusted. I highly recommend checking the inline fuse to the heater if you suspect it is not working. You can find it tucked inside behind the oven. It should be covered in heat shrink. Lastly, there was a rev done that does not appear in the downloadable users manual on Didier's website. It appears the heaters were separated into two groups of two instead of one group of four. Some of the parts were mounted under the oven assembly on the A8 Preheater PCA. If you look at the A4 board from the top, you will see either one or two transistors mounted to the frame (Q4,Q14). If you see only one (Q4), chances are you have the newer rev. I bought the manual recently and just noticed these revs were in my book. I will try and scan the extra info and upload them. Todd
JL
J. L. Trantham
Mon, Dec 30, 2013 12:51 AM

Todd,

Very good information.

Sounds like we are getting close to the correct P/N for the mating plug for
the socket.

My 732A has only a single transistor mounted to the frame from the A4
Regulator Assembly, in the position of Q4, and the resistance is 4230 ohms
after the unit is 'warmed up' and stable.

I recently bought a manual as well but it did not have any 'update sheets'.
I would love to have a copy of your 'update sheets'.

I did a 'clean up' of the update sheets on Didier's site and can send them
to anyone that wants one.

I printed the .PDF from the Fluke site and added the 'cleaned up' version of
the update sheets and it makes a very nice manual.  I would love to add the
other sheets to this.

I also have a 735C that is very similar to the 732A but lot's of mechanical
differences.  It has two transistors on the Regulator Assembly, presumably
A4 and Q4 and Q14.  I do not have a manual for the 735C.

It's thermistor value is in the mid 3K range.  It, too, is very stable and
has many similarities to the 732A but has recently developed a 'new
problem'.  The 'NO CAL' LED comes on after a few days for no apparent
reason.  I need to open it up again and take a look at that circuit to see
what the problem is.

On the 732A, the 'IN CAL' LED is continuously lighted when there has been no
power interruption.  On the 735C, the 'NO CAL' LED illuminates when power is
restored after a power failure.

I replaced the NiCd battery pack and all seems to be well but the 'NO CAL'
LED comes on after having been reset (by inserting a copper wire in the
'adjust' opening).  I wonder if my 'reset' process is the problem.  On the
732A, you connect to the negative terminal then touch the 'pad' inside the
'RESET' opening.  There is no such 'RESET' opening on the 735C.  Just
placing the wire (that I use to adjust the 10 V adjustment) inside the
adjustment opening and aiming to the side would reset the LED.

Happy to help with the scanning of the manual and update sheets if needed.

Thanks again for the info.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of T. Micallef
Sent: Sunday, December 29, 2013 6:05 PM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe,

I have the external battery pack (732A-7003) that I picked up a while ago.
It is indeed the D01 connector. It is printed clearly on the outside edge
along with the three letters that look like ARB.

The company's website states they are now owned by Smiths Connectors and the
connector is part of Hypertac. No list of distributors was found on the
website.

I found a seller Astrex Electronics, but they do not have a price listed for
the female version, just the male. I searched for D01P306FSTAH. This is the
female plug with solder cups.

To date, I have reconditioned three 732A's and they have been operating fine
with little issues, except when I lost power for an extended time and two
batteries in one of the packs died. All four were replaced and they have
been working fine.

The two older 732A's have thermistor values in the 3k-4k range. The newer
unit (based on the newer heater rev) is around 4.25K.

All three had their potentiometers changed on the regulator and
pre-regulator boards. The datecodes on all appeared to be from 1985. The
parts list must have been updated as some of the pots did not have the same
value as the ones installed. I cannot remember which ones were different. I
did not see a mention in the change/errata.

Also, I had to remove the reference ovens to repair broken 10V
potentiometers, as someone had broken one (free spinning) and another unit
appeared to have a noisy wiper so it was replaced as well. The 1V and 1.018V
pots were not changed. The third 10V pot was replaced because I had success
with the other two and it was much easier the third time around. The jumpers
on the all the A7 boards were adjusted.

I highly recommend checking the inline fuse to the heater if you suspect it
is not working. You can find it tucked inside behind the oven. It should be
covered in heat shrink.

Lastly, there was a rev done that does not appear in the downloadable users
manual on Didier's website.  It appears the heaters were separated into two
groups of two instead of one group of four. Some of the parts were mounted
under the oven assembly on the A8 Preheater PCA. If you look at the A4 board
from the top, you will see either one or two transistors mounted to the
frame (Q4,Q14). If you see only one (Q4), chances are you have the newer
rev.

I bought the manual recently and just noticed these revs were in my book.
I will try and scan the extra info and upload them.

Todd


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Todd, Very good information. Sounds like we are getting close to the correct P/N for the mating plug for the socket. My 732A has only a single transistor mounted to the frame from the A4 Regulator Assembly, in the position of Q4, and the resistance is 4230 ohms after the unit is 'warmed up' and stable. I recently bought a manual as well but it did not have any 'update sheets'. I would love to have a copy of your 'update sheets'. I did a 'clean up' of the update sheets on Didier's site and can send them to anyone that wants one. I printed the .PDF from the Fluke site and added the 'cleaned up' version of the update sheets and it makes a very nice manual. I would love to add the other sheets to this. I also have a 735C that is very similar to the 732A but lot's of mechanical differences. It has two transistors on the Regulator Assembly, presumably A4 and Q4 and Q14. I do not have a manual for the 735C. It's thermistor value is in the mid 3K range. It, too, is very stable and has many similarities to the 732A but has recently developed a 'new problem'. The 'NO CAL' LED comes on after a few days for no apparent reason. I need to open it up again and take a look at that circuit to see what the problem is. On the 732A, the 'IN CAL' LED is continuously lighted when there has been no power interruption. On the 735C, the 'NO CAL' LED illuminates when power is restored after a power failure. I replaced the NiCd battery pack and all seems to be well but the 'NO CAL' LED comes on after having been reset (by inserting a copper wire in the 'adjust' opening). I wonder if my 'reset' process is the problem. On the 732A, you connect to the negative terminal then touch the 'pad' inside the 'RESET' opening. There is no such 'RESET' opening on the 735C. Just placing the wire (that I use to adjust the 10 V adjustment) inside the adjustment opening and aiming to the side would reset the LED. Happy to help with the scanning of the manual and update sheets if needed. Thanks again for the info. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of T. Micallef Sent: Sunday, December 29, 2013 6:05 PM To: volt-nuts@febo.com Subject: Re: [volt-nuts] Fluke 732A Questions Joe, I have the external battery pack (732A-7003) that I picked up a while ago. It is indeed the D01 connector. It is printed clearly on the outside edge along with the three letters that look like ARB. The company's website states they are now owned by Smiths Connectors and the connector is part of Hypertac. No list of distributors was found on the website. I found a seller Astrex Electronics, but they do not have a price listed for the female version, just the male. I searched for D01P306FSTAH. This is the female plug with solder cups. To date, I have reconditioned three 732A's and they have been operating fine with little issues, except when I lost power for an extended time and two batteries in one of the packs died. All four were replaced and they have been working fine. The two older 732A's have thermistor values in the 3k-4k range. The newer unit (based on the newer heater rev) is around 4.25K. All three had their potentiometers changed on the regulator and pre-regulator boards. The datecodes on all appeared to be from 1985. The parts list must have been updated as some of the pots did not have the same value as the ones installed. I cannot remember which ones were different. I did not see a mention in the change/errata. Also, I had to remove the reference ovens to repair broken 10V potentiometers, as someone had broken one (free spinning) and another unit appeared to have a noisy wiper so it was replaced as well. The 1V and 1.018V pots were not changed. The third 10V pot was replaced because I had success with the other two and it was much easier the third time around. The jumpers on the all the A7 boards were adjusted. I highly recommend checking the inline fuse to the heater if you suspect it is not working. You can find it tucked inside behind the oven. It should be covered in heat shrink. Lastly, there was a rev done that does not appear in the downloadable users manual on Didier's website. It appears the heaters were separated into two groups of two instead of one group of four. Some of the parts were mounted under the oven assembly on the A8 Preheater PCA. If you look at the A4 board from the top, you will see either one or two transistors mounted to the frame (Q4,Q14). If you see only one (Q4), chances are you have the newer rev. I bought the manual recently and just noticed these revs were in my book. I will try and scan the extra info and upload them. Todd _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
TM
T. Micallef
Mon, Dec 30, 2013 4:03 AM

Joe,

I don't know if it makes any difference but my manual is for the 732A/AN p/n
788414

I found a couple of brochures that I scanned and uploaded too.

I forgot to mention, I have been collecting information regarding the
techniques used to do intercomparisons using scanners like the Data Proof
models. While searching a couple of weeks ago, I came across an old document
from a Measurement Science Conference back in 1983. The document contains an
article regarding the testing of the Fluke 735C. I thought you may have
little to no info regarding your reference and thought I would let you know
about it.

The document is easy to find using google. It is named MSC1983.pdf
I will upload it as well, but I was not sure if Didier wants to track
conference docs.

Todd

Joe, I don't know if it makes any difference but my manual is for the 732A/AN p/n 788414 I found a couple of brochures that I scanned and uploaded too. I forgot to mention, I have been collecting information regarding the techniques used to do intercomparisons using scanners like the Data Proof models. While searching a couple of weeks ago, I came across an old document from a Measurement Science Conference back in 1983. The document contains an article regarding the testing of the Fluke 735C. I thought you may have little to no info regarding your reference and thought I would let you know about it. The document is easy to find using google. It is named MSC1983.pdf I will upload it as well, but I was not sure if Didier wants to track conference docs. Todd
R
R.Phillips
Mon, Dec 30, 2013 10:30 AM

Hi Todd
I would appreciate a copy of your later issue manual for the 732A - please
let me know when you make this available  -  Dider's site ?
Many thanks
Roy

-----Original Message-----
From: T. Micallef
Sent: Monday, December 30, 2013 4:03 AM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe,

I don't know if it makes any difference but my manual is for the 732A/AN p/n
788414

I found a couple of brochures that I scanned and uploaded too.

I forgot to mention, I have been collecting information regarding the
techniques used to do intercomparisons using scanners like the Data Proof
models. While searching a couple of weeks ago, I came across an old document
from a Measurement Science Conference back in 1983. The document contains an
article regarding the testing of the Fluke 735C. I thought you may have
little to no info regarding your reference and thought I would let you know
about it.

The document is easy to find using google. It is named MSC1983.pdf
I will upload it as well, but I was not sure if Didier wants to track
conference docs.

Todd


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi Todd I would appreciate a copy of your later issue manual for the 732A - please let me know when you make this available - Dider's site ? Many thanks Roy -----Original Message----- From: T. Micallef Sent: Monday, December 30, 2013 4:03 AM To: volt-nuts@febo.com Subject: Re: [volt-nuts] Fluke 732A Questions Joe, I don't know if it makes any difference but my manual is for the 732A/AN p/n 788414 I found a couple of brochures that I scanned and uploaded too. I forgot to mention, I have been collecting information regarding the techniques used to do intercomparisons using scanners like the Data Proof models. While searching a couple of weeks ago, I came across an old document from a Measurement Science Conference back in 1983. The document contains an article regarding the testing of the Fluke 735C. I thought you may have little to no info regarding your reference and thought I would let you know about it. The document is easy to find using google. It is named MSC1983.pdf I will upload it as well, but I was not sure if Didier wants to track conference docs. Todd _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
DJ
Didier Juges
Mon, Dec 30, 2013 12:55 PM

I may be unable to move manuals until the end of the week. However, you can always download a manual that was just uploaded by using ftp. Check the Upload button on the Manuals page.

Happy New Year everyone.

Didier KO4BB

"R.Phillips" phill.r1@btinternet.com wrote:

Hi Todd
I would appreciate a copy of your later issue manual for the 732A -
please
let me know when you make this available  -  Dider's site ?
Many thanks
Roy

-----Original Message-----
From: T. Micallef
Sent: Monday, December 30, 2013 4:03 AM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe,

I don't know if it makes any difference but my manual is for the
732A/AN p/n
788414

I found a couple of brochures that I scanned and uploaded too.

I forgot to mention, I have been collecting information regarding the
techniques used to do intercomparisons using scanners like the Data
Proof
models. While searching a couple of weeks ago, I came across an old
document
from a Measurement Science Conference back in 1983. The document
contains an
article regarding the testing of the Fluke 735C. I thought you may have
little to no info regarding your reference and thought I would let you
know
about it.

The document is easy to find using google. It is named MSC1983.pdf
I will upload it as well, but I was not sure if Didier wants to track
conference docs.

Todd


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

--
Sent from my Motorola Droid Razr 4G LTE wireless tracker while I do other things.

I may be unable to move manuals until the end of the week. However, you can always download a manual that was just uploaded by using ftp. Check the Upload button on the Manuals page. Happy New Year everyone. Didier KO4BB "R.Phillips" <phill.r1@btinternet.com> wrote: >Hi Todd >I would appreciate a copy of your later issue manual for the 732A - >please >let me know when you make this available - Dider's site ? >Many thanks >Roy > > >-----Original Message----- >From: T. Micallef >Sent: Monday, December 30, 2013 4:03 AM >To: volt-nuts@febo.com >Subject: Re: [volt-nuts] Fluke 732A Questions > >Joe, > >I don't know if it makes any difference but my manual is for the >732A/AN p/n >788414 > >I found a couple of brochures that I scanned and uploaded too. > >I forgot to mention, I have been collecting information regarding the >techniques used to do intercomparisons using scanners like the Data >Proof >models. While searching a couple of weeks ago, I came across an old >document >from a Measurement Science Conference back in 1983. The document >contains an >article regarding the testing of the Fluke 735C. I thought you may have >little to no info regarding your reference and thought I would let you >know >about it. > >The document is easy to find using google. It is named MSC1983.pdf >I will upload it as well, but I was not sure if Didier wants to track >conference docs. > >Todd > > > > > >_______________________________________________ >volt-nuts mailing list -- volt-nuts@febo.com >To unsubscribe, go to >https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >and follow the instructions there. > > >_______________________________________________ >volt-nuts mailing list -- volt-nuts@febo.com >To unsubscribe, go to >https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >and follow the instructions there. -- Sent from my Motorola Droid Razr 4G LTE wireless tracker while I do other things.
JL
J. L. Trantham
Mon, Dec 30, 2013 2:13 PM

Todd,

I found the MSC1983.PDF document with no trouble.  Thanks for the info.

Can't wait to see the 'change sheets' for the 732A.  Where did you find your
732A manual?  Mine is the one on the Fluke website, P/N 645051, for the
732A, not 732A/AN.  I'll have to search for that.

Happy New Year.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of T. Micallef
Sent: Sunday, December 29, 2013 10:04 PM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe,

I don't know if it makes any difference but my manual is for the 732A/AN p/n
788414

I found a couple of brochures that I scanned and uploaded too.

I forgot to mention, I have been collecting information regarding the
techniques used to do intercomparisons using scanners like the Data Proof
models. While searching a couple of weeks ago, I came across an old document
from a Measurement Science Conference back in 1983. The document contains an
article regarding the testing of the Fluke 735C. I thought you may have
little to no info regarding your reference and thought I would let you know
about it.

The document is easy to find using google. It is named MSC1983.pdf I will
upload it as well, but I was not sure if Didier wants to track conference
docs.

Todd


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Todd, I found the MSC1983.PDF document with no trouble. Thanks for the info. Can't wait to see the 'change sheets' for the 732A. Where did you find your 732A manual? Mine is the one on the Fluke website, P/N 645051, for the 732A, not 732A/AN. I'll have to search for that. Happy New Year. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of T. Micallef Sent: Sunday, December 29, 2013 10:04 PM To: volt-nuts@febo.com Subject: Re: [volt-nuts] Fluke 732A Questions Joe, I don't know if it makes any difference but my manual is for the 732A/AN p/n 788414 I found a couple of brochures that I scanned and uploaded too. I forgot to mention, I have been collecting information regarding the techniques used to do intercomparisons using scanners like the Data Proof models. While searching a couple of weeks ago, I came across an old document from a Measurement Science Conference back in 1983. The document contains an article regarding the testing of the Fluke 735C. I thought you may have little to no info regarding your reference and thought I would let you know about it. The document is easy to find using google. It is named MSC1983.pdf I will upload it as well, but I was not sure if Didier wants to track conference docs. Todd _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
TM
T. Micallef
Mon, Dec 30, 2013 7:05 PM

My manual came from manualsplus. It was in excellent condition compared to
the ones on eBay. You do have to ask them for a picture of every manual
since they can vary in condition.

I have rescanned the whole manual, which I should have done anyway.
I have two missing pages and I am hopeful the seller will scan them and
forward them to me.

Todd

My manual came from manualsplus. It was in excellent condition compared to the ones on eBay. You do have to ask them for a picture of every manual since they can vary in condition. I have rescanned the whole manual, which I should have done anyway. I have two missing pages and I am hopeful the seller will scan them and forward them to me. Todd
BG
Bill Gold
Tue, Dec 31, 2013 1:02 AM

Joe:

My 732A units vary in thermistor resistance from 3.408K ohms to 4.514K

ohms and have been extremely stable to with in +/- 5 ohms for years now.  I
did measure the oven temperature of the one with the 3.408K ohms just to see
what was causing the thermistor resistance to be a little low and came up
with about 47 degrees centigrade, as best as my equipment would allow me to
measure it.  Equipment was a Fluke 80TK thermocouple module, a Fluke 80 PK-1
thermocouple bead and a HP 3456A DVM.  I put shrink wrap around the bead and
then threaded the bead into the temperature oven through one of the
adjustment holes in the front panel and made sure that it was far into the
oven, probably around the middle.  Then I checked another unit that read
4.5K ohms and it read around 43 degrees C so all of that made sense.  I
believe I saw somewhere that the spec was 45 degrees C +/- 2 C.

I did have to change the jumpers on the "Calibration PCB Assembly" on

one of my units so that it would match other 732A units that were "In Cal"
and had certified values.  But that unit has since drifted down at a rate of
around 1.2 ppm per year which is with in specs but a little more than I have
seen from other units.  Recently this unit has suddenly quit drifting down
and seems a lot more stable now.  That was S/N 459xxxx.  I have a S/N
343xxxx which has been proven rock solid for years now at around +/- .3 ppm
and just seems to have "DC noise" stability as it just goes up a little and
then down a little and never needed any adjustment.  I also have a S/N
460xxxx which is extremely stable also.  So I guess that age does seem to
factor into stability as well as how long they have been powered up.  But
the only way to insure that you have a good "volt" is to have at least 4
units and then inter-compare them periodically following NIST (NBS)
Technical Note #430.  While 430 was written for Saturated Standard Cells,
the technique seems to work just fine at 10 volts also.  You can also find
this measurement technique in the 1st Fluke "Calibration - Philosophy in
Practice" book published around 1974.

I gave up on the "unobtainium" connector on the back of the battery pack

very quickly.  I drilled a 1/4 " holes on either side of the "unobtainium"
connector, after removing it, and used two single miniature banana jacks to
allow me to connect to the batteries for the purpose of checking their
performance as described in the 732A manual page 2-9 in manual P/N 788414
May 1986.  I should probably turn in my resignation as a "volt nut" for this
action but it works.  KISS.

Bill

----- Original Message -----
From: "J. L. Trantham" jltran@att.net
To: volt-nuts@febo.com
Sent: Saturday, December 28, 2013 2:51 PM
Subject: [volt-nuts] Fluke 732A Questions

I have reviewed the prior postings on Volt-nuts (a very rewarding review,
BTW)) regarding the 732A and have two questions:

  1.    Has anyone determined the part number, or a source, for the
    

mating

plug to J10, the external power connector for the 732A-7005 battery pack?

I

noted some postings recently about this but did not find a definitive
response regarding the identity of this plug.  The alternative would

appear

to be a complete replacement of J10 and its plug, as long is it all fits

in

the opening in the panel.  I hate to 'bore holes' in vintage equipment.

  1.   The oven thermistor in my 'new to me' unit measures 4229 ohms
    

(+/-

an ohm or two) after the unit has warmed up for a week or so.  I note the
manual refers to 3K to 4K for the value of this thermistor when the unit

is

'stable' as well as other's posting values in the mid 3500's ohms for

their

units.  Should I be concerned?  The unit seems to be stable to within

about

2 uV over about a week (as measured by my 3458A - see below).  Should I

open

the unit and try to measure the oven temperature or just be satisfied that
the unit seems to be working?

I had to replace the four 6V 4AH SLA batteries and they charged up
appropriately as judged by the front panel LED's.  I had to remove the
'jumper' for the '40' option on the A7 board and connect the jumper to the
'20' and '10' options (total of '30') in order to get the unit to adjust

to

10.0000000 VDC on my 'Agilent In Cal'd' 3458A.

The need to change the jumpers, perhaps, could be just an ageing issue or,
on the other hand, a 'temperature' issue with the oven.

Should I open the unit and directly measure the temperature (supposedly
about 48 degrees C) or just be satisfied with what I have?

My recently added 735C also needed moving some 'jumpers' in order to get

it

'on scale', as determined by my 3458A, although it's thermistor measures
about 3330 ohms (after being on for several weeks).

Thanks for everyone's help.

Joe


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.

Joe: My 732A units vary in thermistor resistance from 3.408K ohms to 4.514K ohms and have been extremely stable to with in +/- 5 ohms for years now. I did measure the oven temperature of the one with the 3.408K ohms just to see what was causing the thermistor resistance to be a little low and came up with about 47 degrees centigrade, as best as my equipment would allow me to measure it. Equipment was a Fluke 80TK thermocouple module, a Fluke 80 PK-1 thermocouple bead and a HP 3456A DVM. I put shrink wrap around the bead and then threaded the bead into the temperature oven through one of the adjustment holes in the front panel and made sure that it was far into the oven, probably around the middle. Then I checked another unit that read 4.5K ohms and it read around 43 degrees C so all of that made sense. I believe I saw somewhere that the spec was 45 degrees C +/- 2 C. I did have to change the jumpers on the "Calibration PCB Assembly" on one of my units so that it would match other 732A units that were "In Cal" and had certified values. But that unit has since drifted down at a rate of around 1.2 ppm per year which is with in specs but a little more than I have seen from other units. Recently this unit has suddenly quit drifting down and seems a lot more stable now. That was S/N 459xxxx. I have a S/N 343xxxx which has been proven rock solid for years now at around +/- .3 ppm and just seems to have "DC noise" stability as it just goes up a little and then down a little and never needed any adjustment. I also have a S/N 460xxxx which is extremely stable also. So I guess that age does seem to factor into stability as well as how long they have been powered up. But the only way to insure that you have a good "volt" is to have at least 4 units and then inter-compare them periodically following NIST (NBS) Technical Note #430. While 430 was written for Saturated Standard Cells, the technique seems to work just fine at 10 volts also. You can also find this measurement technique in the 1st Fluke "Calibration - Philosophy in Practice" book published around 1974. I gave up on the "unobtainium" connector on the back of the battery pack very quickly. I drilled a 1/4 " holes on either side of the "unobtainium" connector, after removing it, and used two single miniature banana jacks to allow me to connect to the batteries for the purpose of checking their performance as described in the 732A manual page 2-9 in manual P/N 788414 May 1986. I should probably turn in my resignation as a "volt nut" for this action but it works. KISS. Bill ----- Original Message ----- From: "J. L. Trantham" <jltran@att.net> To: <volt-nuts@febo.com> Sent: Saturday, December 28, 2013 2:51 PM Subject: [volt-nuts] Fluke 732A Questions > I have reviewed the prior postings on Volt-nuts (a very rewarding review, > BTW)) regarding the 732A and have two questions: > > > > 1. Has anyone determined the part number, or a source, for the mating > plug to J10, the external power connector for the 732A-7005 battery pack? I > noted some postings recently about this but did not find a definitive > response regarding the identity of this plug. The alternative would appear > to be a complete replacement of J10 and its plug, as long is it all fits in > the opening in the panel. I hate to 'bore holes' in vintage equipment. > > 2. The oven thermistor in my 'new to me' unit measures 4229 ohms (+/- > an ohm or two) after the unit has warmed up for a week or so. I note the > manual refers to 3K to 4K for the value of this thermistor when the unit is > 'stable' as well as other's posting values in the mid 3500's ohms for their > units. Should I be concerned? The unit seems to be stable to within about > 2 uV over about a week (as measured by my 3458A - see below). Should I open > the unit and try to measure the oven temperature or just be satisfied that > the unit seems to be working? > > > > I had to replace the four 6V 4AH SLA batteries and they charged up > appropriately as judged by the front panel LED's. I had to remove the > 'jumper' for the '40' option on the A7 board and connect the jumper to the > '20' and '10' options (total of '30') in order to get the unit to adjust to > 10.0000000 VDC on my 'Agilent In Cal'd' 3458A. > > > > The need to change the jumpers, perhaps, could be just an ageing issue or, > on the other hand, a 'temperature' issue with the oven. > > > > Should I open the unit and directly measure the temperature (supposedly > about 48 degrees C) or just be satisfied with what I have? > > > > My recently added 735C also needed moving some 'jumpers' in order to get it > 'on scale', as determined by my 3458A, although it's thermistor measures > about 3330 ohms (after being on for several weeks). > > > > Thanks for everyone's help. > > > > Joe > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
JL
J. L. Trantham
Tue, Dec 31, 2013 2:00 AM

I called them this AM and they only had a 'copy' of the manual and not an
original.

If I understand what you are saying, you got a copy as well.

If so, were the 'change sheets' part of the 'copy'?  When I asked about
that, I was told that there were no 'change sheets' in the copy she was
looking at when I called and asked about condition and change sheets.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of T. Micallef
Sent: Monday, December 30, 2013 1:06 PM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Fluke 732A Questions

My manual came from manualsplus. It was in excellent condition compared to
the ones on eBay. You do have to ask them for a picture of every manual
since they can vary in condition.

I have rescanned the whole manual, which I should have done anyway.
I have two missing pages and I am hopeful the seller will scan them and
forward them to me.

Todd


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

I called them this AM and they only had a 'copy' of the manual and not an original. If I understand what you are saying, you got a copy as well. If so, were the 'change sheets' part of the 'copy'? When I asked about that, I was told that there were no 'change sheets' in the copy she was looking at when I called and asked about condition and change sheets. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of T. Micallef Sent: Monday, December 30, 2013 1:06 PM To: volt-nuts@febo.com Subject: Re: [volt-nuts] Fluke 732A Questions My manual came from manualsplus. It was in excellent condition compared to the ones on eBay. You do have to ask them for a picture of every manual since they can vary in condition. I have rescanned the whole manual, which I should have done anyway. I have two missing pages and I am hopeful the seller will scan them and forward them to me. Todd _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
JL
J. L. Trantham
Tue, Dec 31, 2013 2:28 AM

Bill,

Sounds like you are further along in the 'volt-nut' process than I am.  I'll
have to chase down the NIST paper.  I have been very impressed by the
stability of the 732A and the 735C.

Unfortunately, I don't have any documentation on the 735C and only the early
manual for the 732A.

I would be interested in the specifics of the battery check procedure in
your manual.  Unfortunately, my manual does not have a page 2-9.

I note that there are two banana jacks on the front of the battery module,
both black, one connected to ground and the other not connected to anything.
Any idea of what they are for?

I always like to keep equipment as 'original' as possible.  If I can't find
the correct connector, I'll find something similar that will fit in the same
hole and use it instead.  We'll see what I hear from Fluke and Hypertronics.

Also, how do you get your 732A calibrated?  Do the batteries last long
enough to ship overnight to a facility?

Thanks for the info.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Bill Gold
Sent: Monday, December 30, 2013 7:03 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe:

My 732A units vary in thermistor resistance from 3.408K ohms to 4.514K

ohms and have been extremely stable to with in +/- 5 ohms for years now.  I
did measure the oven temperature of the one with the 3.408K ohms just to see
what was causing the thermistor resistance to be a little low and came up
with about 47 degrees centigrade, as best as my equipment would allow me to
measure it.  Equipment was a Fluke 80TK thermocouple module, a Fluke 80 PK-1
thermocouple bead and a HP 3456A DVM.  I put shrink wrap around the bead and
then threaded the bead into the temperature oven through one of the
adjustment holes in the front panel and made sure that it was far into the
oven, probably around the middle.  Then I checked another unit that read
4.5K ohms and it read around 43 degrees C so all of that made sense.  I
believe I saw somewhere that the spec was 45 degrees C +/- 2 C.

I did have to change the jumpers on the "Calibration PCB Assembly" on

one of my units so that it would match other 732A units that were "In Cal"
and had certified values.  But that unit has since drifted down at a rate of
around 1.2 ppm per year which is with in specs but a little more than I have
seen from other units.  Recently this unit has suddenly quit drifting down
and seems a lot more stable now.  That was S/N 459xxxx.  I have a S/N
343xxxx which has been proven rock solid for years now at around +/- .3 ppm
and just seems to have "DC noise" stability as it just goes up a little and
then down a little and never needed any adjustment.  I also have a S/N
460xxxx which is extremely stable also.  So I guess that age does seem to
factor into stability as well as how long they have been powered up.  But
the only way to insure that you have a good "volt" is to have at least 4
units and then inter-compare them periodically following NIST (NBS)
Technical Note #430.  While 430 was written for Saturated Standard Cells,
the technique seems to work just fine at 10 volts also.  You can also find
this measurement technique in the 1st Fluke "Calibration - Philosophy in
Practice" book published around 1974.

I gave up on the "unobtainium" connector on the back of the battery pack

very quickly.  I drilled a 1/4 " holes on either side of the "unobtainium"
connector, after removing it, and used two single miniature banana jacks to
allow me to connect to the batteries for the purpose of checking their
performance as described in the 732A manual page 2-9 in manual P/N 788414
May 1986.  I should probably turn in my resignation as a "volt nut" for this
action but it works.  KISS.

Bill

----- Original Message -----
From: "J. L. Trantham" jltran@att.net
To: volt-nuts@febo.com
Sent: Saturday, December 28, 2013 2:51 PM
Subject: [volt-nuts] Fluke 732A Questions

I have reviewed the prior postings on Volt-nuts (a very rewarding review,
BTW)) regarding the 732A and have two questions:

  1.    Has anyone determined the part number, or a source, for the
    

mating

plug to J10, the external power connector for the 732A-7005 battery pack?

I

noted some postings recently about this but did not find a definitive
response regarding the identity of this plug.  The alternative would

appear

to be a complete replacement of J10 and its plug, as long is it all fits

in

the opening in the panel.  I hate to 'bore holes' in vintage equipment.

  1.   The oven thermistor in my 'new to me' unit measures 4229 ohms
    

(+/-

an ohm or two) after the unit has warmed up for a week or so.  I note the
manual refers to 3K to 4K for the value of this thermistor when the unit

is

'stable' as well as other's posting values in the mid 3500's ohms for

their

units.  Should I be concerned?  The unit seems to be stable to within

about

2 uV over about a week (as measured by my 3458A - see below).  Should I

open

the unit and try to measure the oven temperature or just be satisfied that
the unit seems to be working?

I had to replace the four 6V 4AH SLA batteries and they charged up
appropriately as judged by the front panel LED's.  I had to remove the
'jumper' for the '40' option on the A7 board and connect the jumper to the
'20' and '10' options (total of '30') in order to get the unit to adjust

to

10.0000000 VDC on my 'Agilent In Cal'd' 3458A.

The need to change the jumpers, perhaps, could be just an ageing issue or,
on the other hand, a 'temperature' issue with the oven.

Should I open the unit and directly measure the temperature (supposedly
about 48 degrees C) or just be satisfied with what I have?

My recently added 735C also needed moving some 'jumpers' in order to get

it

'on scale', as determined by my 3458A, although it's thermistor measures
about 3330 ohms (after being on for several weeks).

Thanks for everyone's help.

Joe


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Bill, Sounds like you are further along in the 'volt-nut' process than I am. I'll have to chase down the NIST paper. I have been very impressed by the stability of the 732A and the 735C. Unfortunately, I don't have any documentation on the 735C and only the early manual for the 732A. I would be interested in the specifics of the battery check procedure in your manual. Unfortunately, my manual does not have a page 2-9. I note that there are two banana jacks on the front of the battery module, both black, one connected to ground and the other not connected to anything. Any idea of what they are for? I always like to keep equipment as 'original' as possible. If I can't find the correct connector, I'll find something similar that will fit in the same hole and use it instead. We'll see what I hear from Fluke and Hypertronics. Also, how do you get your 732A calibrated? Do the batteries last long enough to ship overnight to a facility? Thanks for the info. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Bill Gold Sent: Monday, December 30, 2013 7:03 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 732A Questions Joe: My 732A units vary in thermistor resistance from 3.408K ohms to 4.514K ohms and have been extremely stable to with in +/- 5 ohms for years now. I did measure the oven temperature of the one with the 3.408K ohms just to see what was causing the thermistor resistance to be a little low and came up with about 47 degrees centigrade, as best as my equipment would allow me to measure it. Equipment was a Fluke 80TK thermocouple module, a Fluke 80 PK-1 thermocouple bead and a HP 3456A DVM. I put shrink wrap around the bead and then threaded the bead into the temperature oven through one of the adjustment holes in the front panel and made sure that it was far into the oven, probably around the middle. Then I checked another unit that read 4.5K ohms and it read around 43 degrees C so all of that made sense. I believe I saw somewhere that the spec was 45 degrees C +/- 2 C. I did have to change the jumpers on the "Calibration PCB Assembly" on one of my units so that it would match other 732A units that were "In Cal" and had certified values. But that unit has since drifted down at a rate of around 1.2 ppm per year which is with in specs but a little more than I have seen from other units. Recently this unit has suddenly quit drifting down and seems a lot more stable now. That was S/N 459xxxx. I have a S/N 343xxxx which has been proven rock solid for years now at around +/- .3 ppm and just seems to have "DC noise" stability as it just goes up a little and then down a little and never needed any adjustment. I also have a S/N 460xxxx which is extremely stable also. So I guess that age does seem to factor into stability as well as how long they have been powered up. But the only way to insure that you have a good "volt" is to have at least 4 units and then inter-compare them periodically following NIST (NBS) Technical Note #430. While 430 was written for Saturated Standard Cells, the technique seems to work just fine at 10 volts also. You can also find this measurement technique in the 1st Fluke "Calibration - Philosophy in Practice" book published around 1974. I gave up on the "unobtainium" connector on the back of the battery pack very quickly. I drilled a 1/4 " holes on either side of the "unobtainium" connector, after removing it, and used two single miniature banana jacks to allow me to connect to the batteries for the purpose of checking their performance as described in the 732A manual page 2-9 in manual P/N 788414 May 1986. I should probably turn in my resignation as a "volt nut" for this action but it works. KISS. Bill ----- Original Message ----- From: "J. L. Trantham" <jltran@att.net> To: <volt-nuts@febo.com> Sent: Saturday, December 28, 2013 2:51 PM Subject: [volt-nuts] Fluke 732A Questions > I have reviewed the prior postings on Volt-nuts (a very rewarding review, > BTW)) regarding the 732A and have two questions: > > > > 1. Has anyone determined the part number, or a source, for the mating > plug to J10, the external power connector for the 732A-7005 battery pack? I > noted some postings recently about this but did not find a definitive > response regarding the identity of this plug. The alternative would appear > to be a complete replacement of J10 and its plug, as long is it all fits in > the opening in the panel. I hate to 'bore holes' in vintage equipment. > > 2. The oven thermistor in my 'new to me' unit measures 4229 ohms (+/- > an ohm or two) after the unit has warmed up for a week or so. I note the > manual refers to 3K to 4K for the value of this thermistor when the unit is > 'stable' as well as other's posting values in the mid 3500's ohms for their > units. Should I be concerned? The unit seems to be stable to within about > 2 uV over about a week (as measured by my 3458A - see below). Should I open > the unit and try to measure the oven temperature or just be satisfied that > the unit seems to be working? > > > > I had to replace the four 6V 4AH SLA batteries and they charged up > appropriately as judged by the front panel LED's. I had to remove the > 'jumper' for the '40' option on the A7 board and connect the jumper to the > '20' and '10' options (total of '30') in order to get the unit to adjust to > 10.0000000 VDC on my 'Agilent In Cal'd' 3458A. > > > > The need to change the jumpers, perhaps, could be just an ageing issue or, > on the other hand, a 'temperature' issue with the oven. > > > > Should I open the unit and directly measure the temperature (supposedly > about 48 degrees C) or just be satisfied with what I have? > > > > My recently added 735C also needed moving some 'jumpers' in order to get it > 'on scale', as determined by my 3458A, although it's thermistor measures > about 3330 ohms (after being on for several weeks). > > > > Thanks for everyone's help. > > > > Joe > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
TM
T. Micallef
Tue, Dec 31, 2013 5:05 AM

I received the manual with the plastic gray cover just like many of their
newer manuals. It appears to have sat on a shelf with little use. No user
notes have been found. The changes are part of the manual as an additional
section in the back.

I suspect that since it wasn't used much, no one noticed the missing pages.
Even the two brochures were in immaculate condition.

I am guessing that maybe their copies are from my original and are missing
the pages in question (4-9 through 4-12). Since I didn't get a response from
them, I will call in the morning since I am in need of another manual.

Todd

I received the manual with the plastic gray cover just like many of their newer manuals. It appears to have sat on a shelf with little use. No user notes have been found. The changes are part of the manual as an additional section in the back. I suspect that since it wasn't used much, no one noticed the missing pages. Even the two brochures were in immaculate condition. I am guessing that maybe their copies are from my original and are missing the pages in question (4-9 through 4-12). Since I didn't get a response from them, I will call in the morning since I am in need of another manual. Todd
CS
Charles Steinmetz
Tue, Dec 31, 2013 5:22 AM

Joe wrote:

Also, how do you get your 732A calibrated?  Do the batteries last long
enough to ship overnight to a facility?

No -- I assumed that was why you were looking for a battery
connector.  You install an external battery for shipment
overnight.  Fluke can supply a complete external battery with cable
and plug ($375, last I knew), or you can make your own.  I believe
they can also supply a shipping container, which has a dedicated
space for the Fluke external battery.  (If you don't use Fluke's
container, build something very sturdy with wood and padding and with
handles that make it hard not to carry it and set it down right side up.)

I wouldn't have anybody but Fluke calibrate a 732A.  There are 3
levels of calibration, all with data, from about $700 to about $1500
(the last is calibration by Fluke's Primary Standards Lab).

As far as thermistor values go, the relevant question is not what is
the value in absolute terms, but rather what is the value in relation
to the "as manufactured" value, which was provided by Fluke when the
unit shipped.  This had been recorded by the original owners of each
of my units, and they are all within 2 ohms of the original value today.

If you do not know the original value, then your original measurement
is a "going forward" value and it cannot tell you anything about
where you are in relation to  the original manufactured value.  The
fact that you had to change cal jumpers suggests that either the oven
temperature has shifted since manufacture, the reference assembly has
shifted, or both.

Fluke has said various things in various places about how far the
thermistor readings can drift before you have a problem.  +/- 20 ohms
over the unit's lifetime seems to be a good consensus value.  If it
changes fast -- say, 10 ohms or more in a month -- you have a problem.

Best regards,

Charles

Joe wrote: >Also, how do you get your 732A calibrated? Do the batteries last long >enough to ship overnight to a facility? No -- I assumed that was why you were looking for a battery connector. You install an external battery for shipment overnight. Fluke can supply a complete external battery with cable and plug ($375, last I knew), or you can make your own. I believe they can also supply a shipping container, which has a dedicated space for the Fluke external battery. (If you don't use Fluke's container, build something very sturdy with wood and padding and with handles that make it hard not to carry it and set it down right side up.) I wouldn't have anybody but Fluke calibrate a 732A. There are 3 levels of calibration, all with data, from about $700 to about $1500 (the last is calibration by Fluke's Primary Standards Lab). As far as thermistor values go, the relevant question is not what is the value in absolute terms, but rather what is the value in relation to the "as manufactured" value, which was provided by Fluke when the unit shipped. This had been recorded by the original owners of each of my units, and they are all within 2 ohms of the original value today. If you do not know the original value, then your original measurement is a "going forward" value and it cannot tell you anything about where you are in relation to the original manufactured value. The fact that you had to change cal jumpers suggests that either the oven temperature has shifted since manufacture, the reference assembly has shifted, or both. Fluke has said various things in various places about how far the thermistor readings can drift before you have a problem. +/- 20 ohms over the unit's lifetime seems to be a good consensus value. If it changes fast -- say, 10 ohms or more in a month -- you have a problem. Best regards, Charles
JL
J. L. Trantham
Tue, Dec 31, 2013 1:22 PM

Charles,

I would like to find the mating plug for J10 for several reasons, being able
to get it to calibration being just one of them.  However, I was hoping it
would be good enough to make it to calibration on the internal batteries.

I seem to recall reading that the internal batteries were good for a few
hours but nothing like what the 732B would do, which sounded like the 732B
could make the trip on the internal batteries.

Where did you find the resistance of the thermistor 'as shipped'?  I saw
mention of being supplied to the original owner when shipped as well and I
have had the top cover off and the battery module out to replace the
batteries and found nothing to suggest a thermistor value.  I have also
lifted the top layer of insulation off the oven to move the cal-jumpers but
did not see anything that suggested a thermistor value.

My 'metrology level' standards are my HP 3458A's, calibrated by Agilent, and
Solartron (Ametek) 7081's calibrated by Ametek.  I have been thinking of
getting the 732A calibrated by Fluke but I didn't know if they would still
calibrate the unit.  I wonder if they would have the 'as shipped' thermistor
value on file for my specific serial number?

I, too, read about the stability of the thermistor being the most important
measurement.  That was in the 'change sheets' attached to the manual on
Didier's site.

The 'cal-jumper' was at '40' when I got the unit.  Per my 3458A, I could
only get it down from about +220 uV to about +100 uV with the front panel
adjustment.  By moving the jumper to '20', I could only get it up to about
-12 uV below 10 V by turning the adjustment all the way in the opposite
direction.  However, by connecting jumpers for both the '20' and '10', I was
able to get it at 10 V with the front panel adjustment at about center
position.  I have not 'logged' any values yet but every time I go to the
shop, it is about +/- 1 or 2 uV from 10 V.  I do the 'AUTO CAL' on the 3458A
each morning I go into the shop.

I measure the resistance with a Fluke 8050A (simply because it is the
closest DMM to the standard) and it is 4229 to 4230 ohms each day.

Interestingly, when I got the 732A, undisturbed Fluke Calibration stickers
were in place over each of the adjustments on the front panel, on a back
screw on the top cover, and 'half' of a sticker on the back over the power
supply module, as if the other half was lost when someone removed the
battery module.  Somehow, I find that comforting, given that it now seems so
stable, making me think it is 'normal aging' that is responsible for the
values I measured rather than a problem with the oven, etc..  IIRC, the spec
is 6 ppm per year for 10 V, or 60 uV per year.  I'll have to look at the
power supply module to see if I can find some date codes but I suspect the
unit is 25 to 30 years old.  Unfortunately, there were no 'dates' on the
stickers to tell when the unit was last calibrated.

I was thinking of sending the 3458A's and the 732A out for calibration
simultaneously and seeing what I find when they get back.

I'll call Fluke and ask about their Cal services, the plug, and the shipping
case/battery pack issues.

Thanks everyone for very useful information.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Charles Steinmetz
Sent: Monday, December 30, 2013 11:23 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe wrote:

Also, how do you get your 732A calibrated?  Do the batteries last long
enough to ship overnight to a facility?

No -- I assumed that was why you were looking for a battery connector.  You
install an external battery for shipment overnight.  Fluke can supply a
complete external battery with cable and plug ($375, last I knew), or you
can make your own.  I believe they can also supply a shipping container,
which has a dedicated space for the Fluke external battery.  (If you don't
use Fluke's container, build something very sturdy with wood and padding and
with handles that make it hard not to carry it and set it down right side
up.)

I wouldn't have anybody but Fluke calibrate a 732A.  There are 3 levels of
calibration, all with data, from about $700 to about $1500 (the last is
calibration by Fluke's Primary Standards Lab).

As far as thermistor values go, the relevant question is not what is the
value in absolute terms, but rather what is the value in relation to the "as
manufactured" value, which was provided by Fluke when the unit shipped.
This had been recorded by the original owners of each of my units, and they
are all within 2 ohms of the original value today.

If you do not know the original value, then your original measurement is a
"going forward" value and it cannot tell you anything about where you are in
relation to  the original manufactured value.  The fact that you had to
change cal jumpers suggests that either the oven temperature has shifted
since manufacture, the reference assembly has shifted, or both.

Fluke has said various things in various places about how far the thermistor
readings can drift before you have a problem.  +/- 20 ohms over the unit's
lifetime seems to be a good consensus value.  If it changes fast -- say, 10
ohms or more in a month -- you have a problem.

Best regards,

Charles


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Charles, I would like to find the mating plug for J10 for several reasons, being able to get it to calibration being just one of them. However, I was hoping it would be good enough to make it to calibration on the internal batteries. I seem to recall reading that the internal batteries were good for a few hours but nothing like what the 732B would do, which sounded like the 732B could make the trip on the internal batteries. Where did you find the resistance of the thermistor 'as shipped'? I saw mention of being supplied to the original owner when shipped as well and I have had the top cover off and the battery module out to replace the batteries and found nothing to suggest a thermistor value. I have also lifted the top layer of insulation off the oven to move the cal-jumpers but did not see anything that suggested a thermistor value. My 'metrology level' standards are my HP 3458A's, calibrated by Agilent, and Solartron (Ametek) 7081's calibrated by Ametek. I have been thinking of getting the 732A calibrated by Fluke but I didn't know if they would still calibrate the unit. I wonder if they would have the 'as shipped' thermistor value on file for my specific serial number? I, too, read about the stability of the thermistor being the most important measurement. That was in the 'change sheets' attached to the manual on Didier's site. The 'cal-jumper' was at '40' when I got the unit. Per my 3458A, I could only get it down from about +220 uV to about +100 uV with the front panel adjustment. By moving the jumper to '20', I could only get it up to about -12 uV below 10 V by turning the adjustment all the way in the opposite direction. However, by connecting jumpers for both the '20' and '10', I was able to get it at 10 V with the front panel adjustment at about center position. I have not 'logged' any values yet but every time I go to the shop, it is about +/- 1 or 2 uV from 10 V. I do the 'AUTO CAL' on the 3458A each morning I go into the shop. I measure the resistance with a Fluke 8050A (simply because it is the closest DMM to the standard) and it is 4229 to 4230 ohms each day. Interestingly, when I got the 732A, undisturbed Fluke Calibration stickers were in place over each of the adjustments on the front panel, on a back screw on the top cover, and 'half' of a sticker on the back over the power supply module, as if the other half was lost when someone removed the battery module. Somehow, I find that comforting, given that it now seems so stable, making me think it is 'normal aging' that is responsible for the values I measured rather than a problem with the oven, etc.. IIRC, the spec is 6 ppm per year for 10 V, or 60 uV per year. I'll have to look at the power supply module to see if I can find some date codes but I suspect the unit is 25 to 30 years old. Unfortunately, there were no 'dates' on the stickers to tell when the unit was last calibrated. I was thinking of sending the 3458A's and the 732A out for calibration simultaneously and seeing what I find when they get back. I'll call Fluke and ask about their Cal services, the plug, and the shipping case/battery pack issues. Thanks everyone for very useful information. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Charles Steinmetz Sent: Monday, December 30, 2013 11:23 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 732A Questions Joe wrote: >Also, how do you get your 732A calibrated? Do the batteries last long >enough to ship overnight to a facility? No -- I assumed that was why you were looking for a battery connector. You install an external battery for shipment overnight. Fluke can supply a complete external battery with cable and plug ($375, last I knew), or you can make your own. I believe they can also supply a shipping container, which has a dedicated space for the Fluke external battery. (If you don't use Fluke's container, build something very sturdy with wood and padding and with handles that make it hard not to carry it and set it down right side up.) I wouldn't have anybody but Fluke calibrate a 732A. There are 3 levels of calibration, all with data, from about $700 to about $1500 (the last is calibration by Fluke's Primary Standards Lab). As far as thermistor values go, the relevant question is not what is the value in absolute terms, but rather what is the value in relation to the "as manufactured" value, which was provided by Fluke when the unit shipped. This had been recorded by the original owners of each of my units, and they are all within 2 ohms of the original value today. If you do not know the original value, then your original measurement is a "going forward" value and it cannot tell you anything about where you are in relation to the original manufactured value. The fact that you had to change cal jumpers suggests that either the oven temperature has shifted since manufacture, the reference assembly has shifted, or both. Fluke has said various things in various places about how far the thermistor readings can drift before you have a problem. +/- 20 ohms over the unit's lifetime seems to be a good consensus value. If it changes fast -- say, 10 ohms or more in a month -- you have a problem. Best regards, Charles _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
R
R.Phillips
Tue, Dec 31, 2013 1:52 PM

Joe
Reference the thermistor value and its recorded value - I asked Fluke this
question and the answer was, they did supply the individual 732A's value in
written form to the original purchaser, but unfortunately they did not
maintain a separate record.
Roy

-----Original Message-----
From: J. L. Trantham
Sent: Tuesday, December 31, 2013 1:22 PM
To: 'Discussion of precise voltage measurement'
Subject: Re: [volt-nuts] Fluke 732A Questions

Charles,

I would like to find the mating plug for J10 for several reasons, being able
to get it to calibration being just one of them.  However, I was hoping it
would be good enough to make it to calibration on the internal batteries.

I seem to recall reading that the internal batteries were good for a few
hours but nothing like what the 732B would do, which sounded like the 732B
could make the trip on the internal batteries.

Where did you find the resistance of the thermistor 'as shipped'?  I saw
mention of being supplied to the original owner when shipped as well and I
have had the top cover off and the battery module out to replace the
batteries and found nothing to suggest a thermistor value.  I have also
lifted the top layer of insulation off the oven to move the cal-jumpers but
did not see anything that suggested a thermistor value.

My 'metrology level' standards are my HP 3458A's, calibrated by Agilent, and
Solartron (Ametek) 7081's calibrated by Ametek.  I have been thinking of
getting the 732A calibrated by Fluke but I didn't know if they would still
calibrate the unit.  I wonder if they would have the 'as shipped' thermistor
value on file for my specific serial number?

I, too, read about the stability of the thermistor being the most important
measurement.  That was in the 'change sheets' attached to the manual on
Didier's site.

The 'cal-jumper' was at '40' when I got the unit.  Per my 3458A, I could
only get it down from about +220 uV to about +100 uV with the front panel
adjustment.  By moving the jumper to '20', I could only get it up to about
-12 uV below 10 V by turning the adjustment all the way in the opposite
direction.  However, by connecting jumpers for both the '20' and '10', I was
able to get it at 10 V with the front panel adjustment at about center
position.  I have not 'logged' any values yet but every time I go to the
shop, it is about +/- 1 or 2 uV from 10 V.  I do the 'AUTO CAL' on the 3458A
each morning I go into the shop.

I measure the resistance with a Fluke 8050A (simply because it is the
closest DMM to the standard) and it is 4229 to 4230 ohms each day.

Interestingly, when I got the 732A, undisturbed Fluke Calibration stickers
were in place over each of the adjustments on the front panel, on a back
screw on the top cover, and 'half' of a sticker on the back over the power
supply module, as if the other half was lost when someone removed the
battery module.  Somehow, I find that comforting, given that it now seems so
stable, making me think it is 'normal aging' that is responsible for the
values I measured rather than a problem with the oven, etc..  IIRC, the spec
is 6 ppm per year for 10 V, or 60 uV per year.  I'll have to look at the
power supply module to see if I can find some date codes but I suspect the
unit is 25 to 30 years old.  Unfortunately, there were no 'dates' on the
stickers to tell when the unit was last calibrated.

I was thinking of sending the 3458A's and the 732A out for calibration
simultaneously and seeing what I find when they get back.

I'll call Fluke and ask about their Cal services, the plug, and the shipping
case/battery pack issues.

Thanks everyone for very useful information.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Charles Steinmetz
Sent: Monday, December 30, 2013 11:23 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe wrote:

Also, how do you get your 732A calibrated?  Do the batteries last long
enough to ship overnight to a facility?

No -- I assumed that was why you were looking for a battery connector.  You
install an external battery for shipment overnight.  Fluke can supply a
complete external battery with cable and plug ($375, last I knew), or you
can make your own.  I believe they can also supply a shipping container,
which has a dedicated space for the Fluke external battery.  (If you don't
use Fluke's container, build something very sturdy with wood and padding and
with handles that make it hard not to carry it and set it down right side
up.)

I wouldn't have anybody but Fluke calibrate a 732A.  There are 3 levels of
calibration, all with data, from about $700 to about $1500 (the last is
calibration by Fluke's Primary Standards Lab).

As far as thermistor values go, the relevant question is not what is the
value in absolute terms, but rather what is the value in relation to the "as
manufactured" value, which was provided by Fluke when the unit shipped.
This had been recorded by the original owners of each of my units, and they
are all within 2 ohms of the original value today.

If you do not know the original value, then your original measurement is a
"going forward" value and it cannot tell you anything about where you are in
relation to  the original manufactured value.  The fact that you had to
change cal jumpers suggests that either the oven temperature has shifted
since manufacture, the reference assembly has shifted, or both.

Fluke has said various things in various places about how far the thermistor
readings can drift before you have a problem.  +/- 20 ohms over the unit's
lifetime seems to be a good consensus value.  If it changes fast -- say, 10
ohms or more in a month -- you have a problem.

Best regards,

Charles


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Joe Reference the thermistor value and its recorded value - I asked Fluke this question and the answer was, they did supply the individual 732A's value in written form to the original purchaser, but unfortunately they did not maintain a separate record. Roy -----Original Message----- From: J. L. Trantham Sent: Tuesday, December 31, 2013 1:22 PM To: 'Discussion of precise voltage measurement' Subject: Re: [volt-nuts] Fluke 732A Questions Charles, I would like to find the mating plug for J10 for several reasons, being able to get it to calibration being just one of them. However, I was hoping it would be good enough to make it to calibration on the internal batteries. I seem to recall reading that the internal batteries were good for a few hours but nothing like what the 732B would do, which sounded like the 732B could make the trip on the internal batteries. Where did you find the resistance of the thermistor 'as shipped'? I saw mention of being supplied to the original owner when shipped as well and I have had the top cover off and the battery module out to replace the batteries and found nothing to suggest a thermistor value. I have also lifted the top layer of insulation off the oven to move the cal-jumpers but did not see anything that suggested a thermistor value. My 'metrology level' standards are my HP 3458A's, calibrated by Agilent, and Solartron (Ametek) 7081's calibrated by Ametek. I have been thinking of getting the 732A calibrated by Fluke but I didn't know if they would still calibrate the unit. I wonder if they would have the 'as shipped' thermistor value on file for my specific serial number? I, too, read about the stability of the thermistor being the most important measurement. That was in the 'change sheets' attached to the manual on Didier's site. The 'cal-jumper' was at '40' when I got the unit. Per my 3458A, I could only get it down from about +220 uV to about +100 uV with the front panel adjustment. By moving the jumper to '20', I could only get it up to about -12 uV below 10 V by turning the adjustment all the way in the opposite direction. However, by connecting jumpers for both the '20' and '10', I was able to get it at 10 V with the front panel adjustment at about center position. I have not 'logged' any values yet but every time I go to the shop, it is about +/- 1 or 2 uV from 10 V. I do the 'AUTO CAL' on the 3458A each morning I go into the shop. I measure the resistance with a Fluke 8050A (simply because it is the closest DMM to the standard) and it is 4229 to 4230 ohms each day. Interestingly, when I got the 732A, undisturbed Fluke Calibration stickers were in place over each of the adjustments on the front panel, on a back screw on the top cover, and 'half' of a sticker on the back over the power supply module, as if the other half was lost when someone removed the battery module. Somehow, I find that comforting, given that it now seems so stable, making me think it is 'normal aging' that is responsible for the values I measured rather than a problem with the oven, etc.. IIRC, the spec is 6 ppm per year for 10 V, or 60 uV per year. I'll have to look at the power supply module to see if I can find some date codes but I suspect the unit is 25 to 30 years old. Unfortunately, there were no 'dates' on the stickers to tell when the unit was last calibrated. I was thinking of sending the 3458A's and the 732A out for calibration simultaneously and seeing what I find when they get back. I'll call Fluke and ask about their Cal services, the plug, and the shipping case/battery pack issues. Thanks everyone for very useful information. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Charles Steinmetz Sent: Monday, December 30, 2013 11:23 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 732A Questions Joe wrote: >Also, how do you get your 732A calibrated? Do the batteries last long >enough to ship overnight to a facility? No -- I assumed that was why you were looking for a battery connector. You install an external battery for shipment overnight. Fluke can supply a complete external battery with cable and plug ($375, last I knew), or you can make your own. I believe they can also supply a shipping container, which has a dedicated space for the Fluke external battery. (If you don't use Fluke's container, build something very sturdy with wood and padding and with handles that make it hard not to carry it and set it down right side up.) I wouldn't have anybody but Fluke calibrate a 732A. There are 3 levels of calibration, all with data, from about $700 to about $1500 (the last is calibration by Fluke's Primary Standards Lab). As far as thermistor values go, the relevant question is not what is the value in absolute terms, but rather what is the value in relation to the "as manufactured" value, which was provided by Fluke when the unit shipped. This had been recorded by the original owners of each of my units, and they are all within 2 ohms of the original value today. If you do not know the original value, then your original measurement is a "going forward" value and it cannot tell you anything about where you are in relation to the original manufactured value. The fact that you had to change cal jumpers suggests that either the oven temperature has shifted since manufacture, the reference assembly has shifted, or both. Fluke has said various things in various places about how far the thermistor readings can drift before you have a problem. +/- 20 ohms over the unit's lifetime seems to be a good consensus value. If it changes fast -- say, 10 ohms or more in a month -- you have a problem. Best regards, Charles _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
JL
J. L. Trantham
Tue, Dec 31, 2013 2:09 PM

I was afraid of that.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of R.Phillips
Sent: Tuesday, December 31, 2013 7:53 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe
Reference the thermistor value and its recorded value - I asked Fluke this
question and the answer was, they did supply the individual 732A's value in
written form to the original purchaser, but unfortunately they did not
maintain a separate record.
Roy

-----Original Message-----
From: J. L. Trantham
Sent: Tuesday, December 31, 2013 1:22 PM
To: 'Discussion of precise voltage measurement'
Subject: Re: [volt-nuts] Fluke 732A Questions

Charles,

I would like to find the mating plug for J10 for several reasons, being able
to get it to calibration being just one of them.  However, I was hoping it
would be good enough to make it to calibration on the internal batteries.

I seem to recall reading that the internal batteries were good for a few
hours but nothing like what the 732B would do, which sounded like the 732B
could make the trip on the internal batteries.

Where did you find the resistance of the thermistor 'as shipped'?  I saw
mention of being supplied to the original owner when shipped as well and I
have had the top cover off and the battery module out to replace the
batteries and found nothing to suggest a thermistor value.  I have also
lifted the top layer of insulation off the oven to move the cal-jumpers but
did not see anything that suggested a thermistor value.

My 'metrology level' standards are my HP 3458A's, calibrated by Agilent, and
Solartron (Ametek) 7081's calibrated by Ametek.  I have been thinking of
getting the 732A calibrated by Fluke but I didn't know if they would still
calibrate the unit.  I wonder if they would have the 'as shipped' thermistor
value on file for my specific serial number?

I, too, read about the stability of the thermistor being the most important
measurement.  That was in the 'change sheets' attached to the manual on
Didier's site.

The 'cal-jumper' was at '40' when I got the unit.  Per my 3458A, I could
only get it down from about +220 uV to about +100 uV with the front panel
adjustment.  By moving the jumper to '20', I could only get it up to about
-12 uV below 10 V by turning the adjustment all the way in the opposite
direction.  However, by connecting jumpers for both the '20' and '10', I was
able to get it at 10 V with the front panel adjustment at about center
position.  I have not 'logged' any values yet but every time I go to the
shop, it is about +/- 1 or 2 uV from 10 V.  I do the 'AUTO CAL' on the 3458A
each morning I go into the shop.

I measure the resistance with a Fluke 8050A (simply because it is the
closest DMM to the standard) and it is 4229 to 4230 ohms each day.

Interestingly, when I got the 732A, undisturbed Fluke Calibration stickers
were in place over each of the adjustments on the front panel, on a back
screw on the top cover, and 'half' of a sticker on the back over the power
supply module, as if the other half was lost when someone removed the
battery module.  Somehow, I find that comforting, given that it now seems so
stable, making me think it is 'normal aging' that is responsible for the
values I measured rather than a problem with the oven, etc..  IIRC, the spec
is 6 ppm per year for 10 V, or 60 uV per year.  I'll have to look at the
power supply module to see if I can find some date codes but I suspect the
unit is 25 to 30 years old.  Unfortunately, there were no 'dates' on the
stickers to tell when the unit was last calibrated.

I was thinking of sending the 3458A's and the 732A out for calibration
simultaneously and seeing what I find when they get back.

I'll call Fluke and ask about their Cal services, the plug, and the shipping
case/battery pack issues.

Thanks everyone for very useful information.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Charles Steinmetz
Sent: Monday, December 30, 2013 11:23 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe wrote:

Also, how do you get your 732A calibrated?  Do the batteries last long
enough to ship overnight to a facility?

No -- I assumed that was why you were looking for a battery connector.  You
install an external battery for shipment overnight.  Fluke can supply a
complete external battery with cable and plug ($375, last I knew), or you
can make your own.  I believe they can also supply a shipping container,
which has a dedicated space for the Fluke external battery.  (If you don't
use Fluke's container, build something very sturdy with wood and padding and
with handles that make it hard not to carry it and set it down right side
up.)

I wouldn't have anybody but Fluke calibrate a 732A.  There are 3 levels of
calibration, all with data, from about $700 to about $1500 (the last is
calibration by Fluke's Primary Standards Lab).

As far as thermistor values go, the relevant question is not what is the
value in absolute terms, but rather what is the value in relation to the "as
manufactured" value, which was provided by Fluke when the unit shipped.
This had been recorded by the original owners of each of my units, and they
are all within 2 ohms of the original value today.

If you do not know the original value, then your original measurement is a
"going forward" value and it cannot tell you anything about where you are in
relation to  the original manufactured value.  The fact that you had to
change cal jumpers suggests that either the oven temperature has shifted
since manufacture, the reference assembly has shifted, or both.

Fluke has said various things in various places about how far the thermistor
readings can drift before you have a problem.  +/- 20 ohms over the unit's
lifetime seems to be a good consensus value.  If it changes fast -- say, 10
ohms or more in a month -- you have a problem.

Best regards,

Charles


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

I was afraid of that. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of R.Phillips Sent: Tuesday, December 31, 2013 7:53 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 732A Questions Joe Reference the thermistor value and its recorded value - I asked Fluke this question and the answer was, they did supply the individual 732A's value in written form to the original purchaser, but unfortunately they did not maintain a separate record. Roy -----Original Message----- From: J. L. Trantham Sent: Tuesday, December 31, 2013 1:22 PM To: 'Discussion of precise voltage measurement' Subject: Re: [volt-nuts] Fluke 732A Questions Charles, I would like to find the mating plug for J10 for several reasons, being able to get it to calibration being just one of them. However, I was hoping it would be good enough to make it to calibration on the internal batteries. I seem to recall reading that the internal batteries were good for a few hours but nothing like what the 732B would do, which sounded like the 732B could make the trip on the internal batteries. Where did you find the resistance of the thermistor 'as shipped'? I saw mention of being supplied to the original owner when shipped as well and I have had the top cover off and the battery module out to replace the batteries and found nothing to suggest a thermistor value. I have also lifted the top layer of insulation off the oven to move the cal-jumpers but did not see anything that suggested a thermistor value. My 'metrology level' standards are my HP 3458A's, calibrated by Agilent, and Solartron (Ametek) 7081's calibrated by Ametek. I have been thinking of getting the 732A calibrated by Fluke but I didn't know if they would still calibrate the unit. I wonder if they would have the 'as shipped' thermistor value on file for my specific serial number? I, too, read about the stability of the thermistor being the most important measurement. That was in the 'change sheets' attached to the manual on Didier's site. The 'cal-jumper' was at '40' when I got the unit. Per my 3458A, I could only get it down from about +220 uV to about +100 uV with the front panel adjustment. By moving the jumper to '20', I could only get it up to about -12 uV below 10 V by turning the adjustment all the way in the opposite direction. However, by connecting jumpers for both the '20' and '10', I was able to get it at 10 V with the front panel adjustment at about center position. I have not 'logged' any values yet but every time I go to the shop, it is about +/- 1 or 2 uV from 10 V. I do the 'AUTO CAL' on the 3458A each morning I go into the shop. I measure the resistance with a Fluke 8050A (simply because it is the closest DMM to the standard) and it is 4229 to 4230 ohms each day. Interestingly, when I got the 732A, undisturbed Fluke Calibration stickers were in place over each of the adjustments on the front panel, on a back screw on the top cover, and 'half' of a sticker on the back over the power supply module, as if the other half was lost when someone removed the battery module. Somehow, I find that comforting, given that it now seems so stable, making me think it is 'normal aging' that is responsible for the values I measured rather than a problem with the oven, etc.. IIRC, the spec is 6 ppm per year for 10 V, or 60 uV per year. I'll have to look at the power supply module to see if I can find some date codes but I suspect the unit is 25 to 30 years old. Unfortunately, there were no 'dates' on the stickers to tell when the unit was last calibrated. I was thinking of sending the 3458A's and the 732A out for calibration simultaneously and seeing what I find when they get back. I'll call Fluke and ask about their Cal services, the plug, and the shipping case/battery pack issues. Thanks everyone for very useful information. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Charles Steinmetz Sent: Monday, December 30, 2013 11:23 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 732A Questions Joe wrote: >Also, how do you get your 732A calibrated? Do the batteries last long >enough to ship overnight to a facility? No -- I assumed that was why you were looking for a battery connector. You install an external battery for shipment overnight. Fluke can supply a complete external battery with cable and plug ($375, last I knew), or you can make your own. I believe they can also supply a shipping container, which has a dedicated space for the Fluke external battery. (If you don't use Fluke's container, build something very sturdy with wood and padding and with handles that make it hard not to carry it and set it down right side up.) I wouldn't have anybody but Fluke calibrate a 732A. There are 3 levels of calibration, all with data, from about $700 to about $1500 (the last is calibration by Fluke's Primary Standards Lab). As far as thermistor values go, the relevant question is not what is the value in absolute terms, but rather what is the value in relation to the "as manufactured" value, which was provided by Fluke when the unit shipped. This had been recorded by the original owners of each of my units, and they are all within 2 ohms of the original value today. If you do not know the original value, then your original measurement is a "going forward" value and it cannot tell you anything about where you are in relation to the original manufactured value. The fact that you had to change cal jumpers suggests that either the oven temperature has shifted since manufacture, the reference assembly has shifted, or both. Fluke has said various things in various places about how far the thermistor readings can drift before you have a problem. +/- 20 ohms over the unit's lifetime seems to be a good consensus value. If it changes fast -- say, 10 ohms or more in a month -- you have a problem. Best regards, Charles _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.