time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

U-blox teaser

RL
Robert LaJeunesse
Fri, Feb 26, 2021 2:34 PM

FWIW. No detailed content, and a rather quick read. "Five key trends in GPS".

https://www.u-blox.com/en/blogs/insights/five-key-trends-gps

FWIW. No detailed content, and a rather quick read. "Five key trends in GPS". https://www.u-blox.com/en/blogs/insights/five-key-trends-gps
JA
John Ackermann N8UR
Fri, Feb 26, 2021 9:27 PM

It's interesting that they talk about the F9 receivers offering 5 ns
absolute time accuracy.  Does anyone know of tests confirming that, and
what sort of care was required in the setup to get there?

John

On 2/26/21 9:34 AM, Robert LaJeunesse wrote:

FWIW. No detailed content, and a rather quick read. "Five key trends in GPS".

https://www.u-blox.com/en/blogs/insights/five-key-trends-gps


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

It's interesting that they talk about the F9 receivers offering 5 ns absolute time accuracy. Does anyone know of tests confirming that, and what sort of care was required in the setup to get there? John ---- On 2/26/21 9:34 AM, Robert LaJeunesse wrote: > FWIW. No detailed content, and a rather quick read. "Five key trends in GPS". > > https://www.u-blox.com/en/blogs/insights/five-key-trends-gps > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >
BK
Bob kb8tq
Fri, Feb 26, 2021 10:26 PM

Hi

I can’t think of many antennas (multi band or single band) that claim to know their
delay to < 5 ns. Simply having a differential delay spec of < 5 ns is quite good.
Same thing with delay ripple, you see specs out to around 15 ns on a lot of antennas.
None of this is getting you to the actual total delay of the antenna. It’s a pretty good
bet that number is a bit larger than either of these.

Some of the ripple probably comes out in the standard modeling. I’m not sure that
the differential delay is taken out that way. Total delay, not taken out in any obvious
fashion ( at least that I can see). If the F9 has a built in antenna database, that’s not
mentioned in the doc’s. Any benefit from the corrections would have to be part of
post processing.

No, that’s not the same as talking about the F9 it’s self doing X ns, but it would be part
of any practical system trying to get close to 5 ns absolute accuracy.

5 ns relative accuracy between two F9’s? I probably could buy that if the appropriate
one sigma / on a clear day / with the wind in the right direction sort of qualifiers are
attached.

Bob

On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR jra@febo.com wrote:

It's interesting that they talk about the F9 receivers offering 5 ns absolute time accuracy.  Does anyone know of tests confirming that, and what sort of care was required in the setup to get there?

John

On 2/26/21 9:34 AM, Robert LaJeunesse wrote:

FWIW. No detailed content, and a rather quick read. "Five key trends in GPS".
https://www.u-blox.com/en/blogs/insights/five-key-trends-gps


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi I can’t think of many antennas (multi band or single band) that claim to know their delay to < 5 ns. Simply having a *differential* delay spec of < 5 ns is quite good. Same thing with delay ripple, you see specs out to around 15 ns on a lot of antennas. None of this is getting you to the actual total delay of the antenna. It’s a pretty good bet that number is a bit larger than either of these. Some of the ripple probably comes out in the standard modeling. I’m not sure that the differential delay is taken out that way. Total delay, not taken out in any obvious fashion ( at least that I can see). If the F9 has a built in antenna database, that’s not mentioned in the doc’s. Any benefit from the corrections would have to be part of post processing. No, that’s not the same as talking about the F9 it’s self doing X ns, but it would be part of any practical system trying to get close to 5 ns absolute accuracy. 5 ns *relative* accuracy between two F9’s? I probably could buy that if the appropriate one sigma / on a clear day / with the wind in the right direction sort of qualifiers are attached. Bob > On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR <jra@febo.com> wrote: > > It's interesting that they talk about the F9 receivers offering 5 ns absolute time accuracy. Does anyone know of tests confirming that, and what sort of care was required in the setup to get there? > > John > ---- > > On 2/26/21 9:34 AM, Robert LaJeunesse wrote: >> FWIW. No detailed content, and a rather quick read. "Five key trends in GPS". >> https://www.u-blox.com/en/blogs/insights/five-key-trends-gps >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
EB
ed breya
Fri, Feb 26, 2021 10:37 PM

John, if you look back at the recent GPSDO discussion initiated by
"dandober" in hpaligent keysight group, you'll see a post by Leo Bodnar,
linking a Ublox model here

https://www.u-blox.com/sites/default/files/products/documents/LEA-M8F_DataSheet_%28UBX-14001772%29.pdf

I looked it up, and I recall the PPS timing spec is +/- 5 nSec. After I
studied it some, I replied back

"Thanks Leo, that's a great example of what I've been picturing. This is
from the data sheet, regarding the PPS output:

"After an initial phase of acquisition the time-pulse becomes
essentially jitter-free, generated coherently from the built-in
reference oscillator and guaranteeing an exact number of reference
frequency cycles between each time-pulse."

It looks like this is an implementation of what we've talked about here
before a few times, for saw-tooth error reduction. BTW I didn't call it
"1PPS" above, because it can be programmed for different rates - I don't
recall what they are.

BTW, did dandober ever sign up here?

Ed

John, if you look back at the recent GPSDO discussion initiated by "dandober" in hpaligent keysight group, you'll see a post by Leo Bodnar, linking a Ublox model here https://www.u-blox.com/sites/default/files/products/documents/LEA-M8F_DataSheet_%28UBX-14001772%29.pdf I looked it up, and I recall the PPS timing spec is +/- 5 nSec. After I studied it some, I replied back "Thanks Leo, that's a great example of what I've been picturing. This is from the data sheet, regarding the PPS output: "After an initial phase of acquisition the time-pulse becomes essentially jitter-free, generated coherently from the built-in reference oscillator and guaranteeing an exact number of reference frequency cycles between each time-pulse." It looks like this is an implementation of what we've talked about here before a few times, for saw-tooth error reduction. BTW I didn't call it "1PPS" above, because it can be programmed for different rates - I don't recall what they are. BTW, did dandober ever sign up here? Ed
BK
Bob kb8tq
Fri, Feb 26, 2021 11:55 PM

Hi

Would not “absolute” timing be referenced to UTC? (or something similar)

Bob

On Feb 26, 2021, at 5:37 PM, ed breya eb@telight.com wrote:

John, if you look back at the recent GPSDO discussion initiated by "dandober" in hpaligent keysight group, you'll see a post by Leo Bodnar, linking a Ublox model here

https://www.u-blox.com/sites/default/files/products/documents/LEA-M8F_DataSheet_%28UBX-14001772%29.pdf

I looked it up, and I recall the PPS timing spec is +/- 5 nSec. After I studied it some, I replied back

"Thanks Leo, that's a great example of what I've been picturing. This is from the data sheet, regarding the PPS output:

"After an initial phase of acquisition the time-pulse becomes essentially jitter-free, generated coherently from the built-in reference oscillator and guaranteeing an exact number of reference frequency cycles between each time-pulse."

It looks like this is an implementation of what we've talked about here before a few times, for saw-tooth error reduction. BTW I didn't call it "1PPS" above, because it can be programmed for different rates - I don't recall what they are.

BTW, did dandober ever sign up here?

Ed


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi Would not “absolute” timing be referenced to UTC? (or something similar) Bob > On Feb 26, 2021, at 5:37 PM, ed breya <eb@telight.com> wrote: > > John, if you look back at the recent GPSDO discussion initiated by "dandober" in hpaligent keysight group, you'll see a post by Leo Bodnar, linking a Ublox model here > > https://www.u-blox.com/sites/default/files/products/documents/LEA-M8F_DataSheet_%28UBX-14001772%29.pdf > > I looked it up, and I recall the PPS timing spec is +/- 5 nSec. After I studied it some, I replied back > > "Thanks Leo, that's a great example of what I've been picturing. This is from the data sheet, regarding the PPS output: > > "After an initial phase of acquisition the time-pulse becomes essentially jitter-free, generated coherently from the built-in reference oscillator and guaranteeing an exact number of reference frequency cycles between each time-pulse." > > It looks like this is an implementation of what we've talked about here before a few times, for saw-tooth error reduction. BTW I didn't call it "1PPS" above, because it can be programmed for different rates - I don't recall what they are. > > BTW, did dandober ever sign up here? > > Ed > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
JA
John Ackermann N8UR
Sat, Feb 27, 2021 12:00 AM

They're claiming "even better than" 5 ns for relative time, which given
the 4 ns jitter seems at least sort-of reasonable.  But until someone
shows me otherwise, I'm still thinking that getting better than 25 ns
absolute accuracy is a pretty good day's work.

John

On 2/26/21 5:26 PM, Bob kb8tq wrote:

Hi

I can’t think of many antennas (multi band or single band) that claim to know their
delay to < 5 ns. Simply having a differential delay spec of < 5 ns is quite good.
Same thing with delay ripple, you see specs out to around 15 ns on a lot of antennas.
None of this is getting you to the actual total delay of the antenna. It’s a pretty good
bet that number is a bit larger than either of these.

Some of the ripple probably comes out in the standard modeling. I’m not sure that
the differential delay is taken out that way. Total delay, not taken out in any obvious
fashion ( at least that I can see). If the F9 has a built in antenna database, that’s not
mentioned in the doc’s. Any benefit from the corrections would have to be part of
post processing.

No, that’s not the same as talking about the F9 it’s self doing X ns, but it would be part
of any practical system trying to get close to 5 ns absolute accuracy.

5 ns relative accuracy between two F9’s? I probably could buy that if the appropriate
one sigma / on a clear day / with the wind in the right direction sort of qualifiers are
attached.

Bob

On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR jra@febo.com wrote:

It's interesting that they talk about the F9 receivers offering 5 ns absolute time accuracy.  Does anyone know of tests confirming that, and what sort of care was required in the setup to get there?

John

On 2/26/21 9:34 AM, Robert LaJeunesse wrote:

FWIW. No detailed content, and a rather quick read. "Five key trends in GPS".
https://www.u-blox.com/en/blogs/insights/five-key-trends-gps


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

They're claiming "even better than" 5 ns for relative time, which given the 4 ns jitter seems at least sort-of reasonable. But until someone shows me otherwise, I'm still thinking that getting better than 25 ns absolute accuracy is a pretty good day's work. John ---- On 2/26/21 5:26 PM, Bob kb8tq wrote: > Hi > > I can’t think of many antennas (multi band or single band) that claim to know their > delay to < 5 ns. Simply having a *differential* delay spec of < 5 ns is quite good. > Same thing with delay ripple, you see specs out to around 15 ns on a lot of antennas. > None of this is getting you to the actual total delay of the antenna. It’s a pretty good > bet that number is a bit larger than either of these. > > Some of the ripple probably comes out in the standard modeling. I’m not sure that > the differential delay is taken out that way. Total delay, not taken out in any obvious > fashion ( at least that I can see). If the F9 has a built in antenna database, that’s not > mentioned in the doc’s. Any benefit from the corrections would have to be part of > post processing. > > No, that’s not the same as talking about the F9 it’s self doing X ns, but it would be part > of any practical system trying to get close to 5 ns absolute accuracy. > > 5 ns *relative* accuracy between two F9’s? I probably could buy that if the appropriate > one sigma / on a clear day / with the wind in the right direction sort of qualifiers are > attached. > > Bob > >> On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR <jra@febo.com> wrote: >> >> It's interesting that they talk about the F9 receivers offering 5 ns absolute time accuracy. Does anyone know of tests confirming that, and what sort of care was required in the setup to get there? >> >> John >> ---- >> >> On 2/26/21 9:34 AM, Robert LaJeunesse wrote: >>> FWIW. No detailed content, and a rather quick read. "Five key trends in GPS". >>> https://www.u-blox.com/en/blogs/insights/five-key-trends-gps >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@lists.febo.com >>> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >
JA
John Ackermann N8UR
Sat, Feb 27, 2021 12:31 AM

You have to be careful comparing the LEA-M8F with other GPS units.  It
does have low jitter because the "TIMEPULSE" signal is derived from the
TCXO which is locked to the GPS time mark.

But their claim of "essentially jitter free" depends on your definition
of "essentially" -- attached is a comparison of LEA-M8F PPS jitter vs.
NEO-M8T and ZED-F9T raw, and NEO-M8T and ZED-F9T sawtooth corrected PPS.
The M8F is definitely better than the raw M8T and even the raw F9T,
but the corrected M8T and F9T are both much better than the M8F.  And
the M8F does not have sawtooth correction available.

So, it's a neat implementation and has some applications (basically as a
modest performance 30.72 MHz GPSDO), but TANSTAAFL.

John

On 2/26/21 5:37 PM, ed breya wrote:

John, if you look back at the recent GPSDO discussion initiated by
"dandober" in hpaligent keysight group, you'll see a post by Leo Bodnar,
linking a Ublox model here

https://www.u-blox.com/sites/default/files/products/documents/LEA-M8F_DataSheet_%28UBX-14001772%29.pdf

I looked it up, and I recall the PPS timing spec is +/- 5 nSec. After I
studied it some, I replied back

"Thanks Leo, that's a great example of what I've been picturing. This is
from the data sheet, regarding the PPS output:

"After an initial phase of acquisition the time-pulse becomes
essentially jitter-free, generated coherently from the built-in
reference oscillator and guaranteeing an exact number of reference
frequency cycles between each time-pulse."

It looks like this is an implementation of what we've talked about here
before a few times, for saw-tooth error reduction. BTW I didn't call it
"1PPS" above, because it can be programmed for different rates - I don't
recall what they are.

BTW, did dandober ever sign up here?

Ed


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

You have to be careful comparing the LEA-M8F with other GPS units. It does have low jitter because the "TIMEPULSE" signal is derived from the TCXO which is locked to the GPS time mark. But their claim of "essentially jitter free" depends on your definition of "essentially" -- attached is a comparison of LEA-M8F PPS jitter vs. NEO-M8T and ZED-F9T raw, and NEO-M8T and ZED-F9T sawtooth corrected PPS. The M8F is definitely better than the raw M8T and even the raw F9T, but the corrected M8T and F9T are both much better than the M8F. And the M8F does *not* have sawtooth correction available. So, it's a neat implementation and has some applications (basically as a modest performance 30.72 MHz GPSDO), but TANSTAAFL. John ---- On 2/26/21 5:37 PM, ed breya wrote: > John, if you look back at the recent GPSDO discussion initiated by > "dandober" in hpaligent keysight group, you'll see a post by Leo Bodnar, > linking a Ublox model here > > https://www.u-blox.com/sites/default/files/products/documents/LEA-M8F_DataSheet_%28UBX-14001772%29.pdf > > > I looked it up, and I recall the PPS timing spec is +/- 5 nSec. After I > studied it some, I replied back > > "Thanks Leo, that's a great example of what I've been picturing. This is > from the data sheet, regarding the PPS output: > > "After an initial phase of acquisition the time-pulse becomes > essentially jitter-free, generated coherently from the built-in > reference oscillator and guaranteeing an exact number of reference > frequency cycles between each time-pulse." > > It looks like this is an implementation of what we've talked about here > before a few times, for saw-tooth error reduction. BTW I didn't call it > "1PPS" above, because it can be programmed for different rates - I don't > recall what they are. > > BTW, did dandober ever sign up here? > > Ed > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
JA
John Ackermann
Sat, Feb 27, 2021 1:01 AM

That's how I'd interpret it.

⁣Get BlueMail for Android ​

On Feb 26, 2021, 7:42 PM, at 7:42 PM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

Would not “absolute” timing be referenced to UTC? (or something
similar)

Bob

On Feb 26, 2021, at 5:37 PM, ed breya eb@telight.com wrote:

John, if you look back at the recent GPSDO discussion initiated by

"dandober" in hpaligent keysight group, you'll see a post by Leo
Bodnar, linking a Ublox model here

I looked it up, and I recall the PPS timing spec is +/- 5 nSec. After

I studied it some, I replied back

"Thanks Leo, that's a great example of what I've been picturing. This

is from the data sheet, regarding the PPS output:

"After an initial phase of acquisition the time-pulse becomes

essentially jitter-free, generated coherently from the built-in
reference oscillator and guaranteeing an exact number of reference
frequency cycles between each time-pulse."

It looks like this is an implementation of what we've talked about

here before a few times, for saw-tooth error reduction. BTW I didn't
call it "1PPS" above, because it can be programmed for different rates

  • I don't recall what they are.

BTW, did dandober ever sign up here?

Ed


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

That's how I'd interpret it. ⁣Get BlueMail for Android ​ On Feb 26, 2021, 7:42 PM, at 7:42 PM, Bob kb8tq <kb8tq@n1k.org> wrote: >Hi > >Would not “absolute” timing be referenced to UTC? (or something >similar) > >Bob > >> On Feb 26, 2021, at 5:37 PM, ed breya <eb@telight.com> wrote: >> >> John, if you look back at the recent GPSDO discussion initiated by >"dandober" in hpaligent keysight group, you'll see a post by Leo >Bodnar, linking a Ublox model here >> >> >https://www.u-blox.com/sites/default/files/products/documents/LEA-M8F_DataSheet_%28UBX-14001772%29.pdf >> >> I looked it up, and I recall the PPS timing spec is +/- 5 nSec. After >I studied it some, I replied back >> >> "Thanks Leo, that's a great example of what I've been picturing. This >is from the data sheet, regarding the PPS output: >> >> "After an initial phase of acquisition the time-pulse becomes >essentially jitter-free, generated coherently from the built-in >reference oscillator and guaranteeing an exact number of reference >frequency cycles between each time-pulse." >> >> It looks like this is an implementation of what we've talked about >here before a few times, for saw-tooth error reduction. BTW I didn't >call it "1PPS" above, because it can be programmed for different rates >- I don't recall what they are. >> >> BTW, did dandober ever sign up here? >> >> Ed >> >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to >http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. > > >_______________________________________________ >time-nuts mailing list -- time-nuts@lists.febo.com >To unsubscribe, go to >http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >and follow the instructions there.
MW
Michael Wouters
Sat, Feb 27, 2021 1:02 AM

Typical L1 antenna delays range from 20 to 70 ns.
I know of only one antenna for which a delay is given by the vendor and the
technique used was just to measure the electronic delay ie by injecting a
signal into the circuit. To do it properly, you need a setup in a microwave
anechoic chamber with transmitting antenna etc. The practical difference
may be small though, 1 or 2 ns ( sample of one antenna!).

Cheers
Michael

On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR jra@febo.com wrote:

They're claiming "even better than" 5 ns for relative time, which given
the 4 ns jitter seems at least sort-of reasonable.  But until someone
shows me otherwise, I'm still thinking that getting better than 25 ns
absolute accuracy is a pretty good day's work.

John

On 2/26/21 5:26 PM, Bob kb8tq wrote:

Hi

I can’t think of many antennas (multi band or single band) that claim to

know their

delay to < 5 ns. Simply having a differential delay spec of < 5 ns is

quite good.

Same thing with delay ripple, you see specs out to around 15 ns on a lot

of antennas.

None of this is getting you to the actual total delay of the antenna.

It’s a pretty good

bet that number is a bit larger than either of these.

Some of the ripple probably comes out in the standard modeling. I’m not

sure that

the differential delay is taken out that way. Total delay, not taken out

in any obvious

fashion ( at least that I can see). If the F9 has a built in antenna

database, that’s not

mentioned in the doc’s. Any benefit from the corrections would have to

be part of

post processing.

No, that’s not the same as talking about the F9 it’s self doing X ns,

but it would be part

of any practical system trying to get close to 5 ns absolute accuracy.

5 ns relative accuracy between two F9’s? I probably could buy that if

the appropriate

one sigma / on a clear day / with the wind in the right direction sort

of qualifiers are

attached.

Bob

On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR jra@febo.com wrote:

It's interesting that they talk about the F9 receivers offering 5 ns

absolute time accuracy.  Does anyone know of tests confirming that, and
what sort of care was required in the setup to get there?

John

On 2/26/21 9:34 AM, Robert LaJeunesse wrote:

FWIW. No detailed content, and a rather quick read. "Five key trends

in GPS".

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Typical L1 antenna delays range from 20 to 70 ns. I know of only one antenna for which a delay is given by the vendor and the technique used was just to measure the electronic delay ie by injecting a signal into the circuit. To do it properly, you need a setup in a microwave anechoic chamber with transmitting antenna etc. The practical difference may be small though, 1 or 2 ns ( sample of one antenna!). Cheers Michael On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR <jra@febo.com> wrote: > They're claiming "even better than" 5 ns for relative time, which given > the 4 ns jitter seems at least sort-of reasonable. But until someone > shows me otherwise, I'm still thinking that getting better than 25 ns > absolute accuracy is a pretty good day's work. > > John > ---- > > On 2/26/21 5:26 PM, Bob kb8tq wrote: > > Hi > > > > I can’t think of many antennas (multi band or single band) that claim to > know their > > delay to < 5 ns. Simply having a *differential* delay spec of < 5 ns is > quite good. > > Same thing with delay ripple, you see specs out to around 15 ns on a lot > of antennas. > > None of this is getting you to the actual total delay of the antenna. > It’s a pretty good > > bet that number is a bit larger than either of these. > > > > Some of the ripple probably comes out in the standard modeling. I’m not > sure that > > the differential delay is taken out that way. Total delay, not taken out > in any obvious > > fashion ( at least that I can see). If the F9 has a built in antenna > database, that’s not > > mentioned in the doc’s. Any benefit from the corrections would have to > be part of > > post processing. > > > > No, that’s not the same as talking about the F9 it’s self doing X ns, > but it would be part > > of any practical system trying to get close to 5 ns absolute accuracy. > > > > 5 ns *relative* accuracy between two F9’s? I probably could buy that if > the appropriate > > one sigma / on a clear day / with the wind in the right direction sort > of qualifiers are > > attached. > > > > Bob > > > >> On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR <jra@febo.com> wrote: > >> > >> It's interesting that they talk about the F9 receivers offering 5 ns > absolute time accuracy. Does anyone know of tests confirming that, and > what sort of care was required in the setup to get there? > >> > >> John > >> ---- > >> > >> On 2/26/21 9:34 AM, Robert LaJeunesse wrote: > >>> FWIW. No detailed content, and a rather quick read. "Five key trends > in GPS". > >>> https://www.u-blox.com/en/blogs/insights/five-key-trends-gps > >>> _______________________________________________ > >>> time-nuts mailing list -- time-nuts@lists.febo.com > >>> To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>> and follow the instructions there. > >> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@lists.febo.com > >> To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >> and follow the instructions there. > > > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@lists.febo.com > > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > > and follow the instructions there. > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >
LJ
Lux, Jim
Sat, Feb 27, 2021 1:06 AM

On 2/26/21 4:00 PM, John Ackermann N8UR wrote:

They're claiming "even better than" 5 ns for relative time, which
given the 4 ns jitter seems at least sort-of reasonable.  But until
someone shows me otherwise, I'm still thinking that getting better
than 25 ns absolute accuracy is a pretty good day's work.

John

And if the jitter is uniformly distributed (i.e. 1 clock cycle of some
sort) - then the sd is 1/sqrt(12) of the extremes, so a "max
uncertainty" of 5ns, is more like a 1.5 ns 1 sigma.

On 2/26/21 4:00 PM, John Ackermann N8UR wrote: > They're claiming "even better than" 5 ns for relative time, which > given the 4 ns jitter seems at least sort-of reasonable.  But until > someone shows me otherwise, I'm still thinking that getting better > than 25 ns absolute accuracy is a pretty good day's work. > > John > ---- And if the jitter is uniformly distributed (i.e. 1 clock cycle of some sort) - then the sd is 1/sqrt(12) of the extremes, so a "max uncertainty" of 5ns, is more like a 1.5 ns 1 sigma.