HP 5065A repair

MH
Matt Huszagh
Sun, Nov 5, 2023 5:07 PM

Hi,

I have an HP 5065A received in unknown condition. After performing some
pre-emptive maintenance (basically Corby's recommendations here:
https://www.eevblog.com/forum/metrology/inspecting-aligning-and-testing-an-hp-5065a-rubidium/
and the first aid advice from Luciano here:
http://www.timeok.it/hp5065a-corner-3/) I turned the unit on. +/- 20V
supplies look good and cell and lamp ovens drop down from pegged in
under 45 min to about 32 each.

However, the 2nd harmonic reading is 0 and of course the unit does not
lock.

This unit was left unpowered for some time (I got the sense it was at
least several years) before I received it, and therefore cell flooding
is likely. However, I have noticed a few other discrepencies.

The first is that the A3 TP2 AGC DC voltage measures 71 mV, instead of
the expected >4.5 V. I also checked the 60 MHz signal into 50 ohms and
see about 30 mW. One place in the manual says this should be >1V (which
it is), but another place says 300 mW. Is loading this with 50 ohms
correct? If so, maybe 300 mW is a typo and it should read 30 mW? In any
event the AGC DC voltage is wrong and I suppose needs to be addressed.

The second issue, and probably a more serious one, is that the PHOTO I
meter reading is 20 (less than the specified 25) and the actual current
measured out of A12 is -10 uA (should be 50 uA or greater). There are
two issues here: (1) the actual current disagrees with the meter
indication and (2) the current is low. Is the low current possibly
explained by cell flooding and or incorrect/insufficient 60 MHz RF
signal application?

Thoughts on next steps? I believe I need to address both issues
above. Addressing cell flooding is straightforward, but will take some
time. The troubleshooting section for A3 says to perform the loop
adjustment procedure if the AGC DC voltage is wrong. Which should I do
first? Or, should I start with something else?

Thanks!
Matt

Hi, I have an HP 5065A received in unknown condition. After performing some pre-emptive maintenance (basically Corby's recommendations here: https://www.eevblog.com/forum/metrology/inspecting-aligning-and-testing-an-hp-5065a-rubidium/ and the first aid advice from Luciano here: http://www.timeok.it/hp5065a-corner-3/) I turned the unit on. +/- 20V supplies look good and cell and lamp ovens drop down from pegged in under 45 min to about 32 each. However, the 2nd harmonic reading is 0 and of course the unit does not lock. This unit was left unpowered for some time (I got the sense it was at least several years) before I received it, and therefore cell flooding is likely. However, I have noticed a few other discrepencies. The first is that the A3 TP2 AGC DC voltage measures 71 mV, instead of the expected >4.5 V. I also checked the 60 MHz signal into 50 ohms and see about 30 mW. One place in the manual says this should be >1V (which it is), but another place says 300 mW. Is loading this with 50 ohms correct? If so, maybe 300 mW is a typo and it should read 30 mW? In any event the AGC DC voltage is wrong and I suppose needs to be addressed. The second issue, and probably a more serious one, is that the PHOTO I meter reading is 20 (less than the specified 25) and the actual current measured out of A12 is -10 uA (should be 50 uA or greater). There are two issues here: (1) the actual current disagrees with the meter indication and (2) the current is low. Is the low current possibly explained by cell flooding and or incorrect/insufficient 60 MHz RF signal application? Thoughts on next steps? I believe I need to address both issues above. Addressing cell flooding is straightforward, but will take some time. The troubleshooting section for A3 says to perform the loop adjustment procedure if the AGC DC voltage is wrong. Which should I do first? Or, should I start with something else? Thanks! Matt
MH
Matt Huszagh
Sun, Nov 5, 2023 5:17 PM

There are a few other things I wanted to mention.

I checked the DC voltages around the cell and lamp heaters. J15P3 (cell
heater) measured 15.4V and J15P5 measured 14.4V. P4 measured 28.6V. The
specified nominal voltages for P3 and P5 are 18V, so these are a bit
off, but since the heater currents look ok as measured by the front
panel meter, I got the sense that this is probably ok. Thoughts?

Also, as part of the pre-emptive maintenance I replaced the lamp
assembly film resistor and, in the process, disassembled part of A12
including the lamp assembly. Everything looked fine to my eye and I have
pictures I can share if desired. In particular, the lamp assembly was
not charred.

Matt

Matt Huszagh huszaghmatt@gmail.com writes:

Hi,

I have an HP 5065A received in unknown condition. After performing some
pre-emptive maintenance (basically Corby's recommendations here:
https://www.eevblog.com/forum/metrology/inspecting-aligning-and-testing-an-hp-5065a-rubidium/
and the first aid advice from Luciano here:
http://www.timeok.it/hp5065a-corner-3/) I turned the unit on. +/- 20V
supplies look good and cell and lamp ovens drop down from pegged in
under 45 min to about 32 each.

However, the 2nd harmonic reading is 0 and of course the unit does not
lock.

This unit was left unpowered for some time (I got the sense it was at
least several years) before I received it, and therefore cell flooding
is likely. However, I have noticed a few other discrepencies.

The first is that the A3 TP2 AGC DC voltage measures 71 mV, instead of
the expected >4.5 V. I also checked the 60 MHz signal into 50 ohms and
see about 30 mW. One place in the manual says this should be >1V (which
it is), but another place says 300 mW. Is loading this with 50 ohms
correct? If so, maybe 300 mW is a typo and it should read 30 mW? In any
event the AGC DC voltage is wrong and I suppose needs to be addressed.

The second issue, and probably a more serious one, is that the PHOTO I
meter reading is 20 (less than the specified 25) and the actual current
measured out of A12 is -10 uA (should be 50 uA or greater). There are
two issues here: (1) the actual current disagrees with the meter
indication and (2) the current is low. Is the low current possibly
explained by cell flooding and or incorrect/insufficient 60 MHz RF
signal application?

Thoughts on next steps? I believe I need to address both issues
above. Addressing cell flooding is straightforward, but will take some
time. The troubleshooting section for A3 says to perform the loop
adjustment procedure if the AGC DC voltage is wrong. Which should I do
first? Or, should I start with something else?

Thanks!
Matt

There are a few other things I wanted to mention. I checked the DC voltages around the cell and lamp heaters. J15P3 (cell heater) measured 15.4V and J15P5 measured 14.4V. P4 measured 28.6V. The specified nominal voltages for P3 and P5 are 18V, so these are a bit off, but since the heater currents look ok as measured by the front panel meter, I got the sense that this is probably ok. Thoughts? Also, as part of the pre-emptive maintenance I replaced the lamp assembly film resistor and, in the process, disassembled part of A12 including the lamp assembly. Everything looked fine to my eye and I have pictures I can share if desired. In particular, the lamp assembly was not charred. Matt Matt Huszagh <huszaghmatt@gmail.com> writes: > Hi, > > I have an HP 5065A received in unknown condition. After performing some > pre-emptive maintenance (basically Corby's recommendations here: > https://www.eevblog.com/forum/metrology/inspecting-aligning-and-testing-an-hp-5065a-rubidium/ > and the first aid advice from Luciano here: > http://www.timeok.it/hp5065a-corner-3/) I turned the unit on. +/- 20V > supplies look good and cell and lamp ovens drop down from pegged in > under 45 min to about 32 each. > > However, the 2nd harmonic reading is 0 and of course the unit does not > lock. > > This unit was left unpowered for some time (I got the sense it was at > least several years) before I received it, and therefore cell flooding > is likely. However, I have noticed a few other discrepencies. > > The first is that the A3 TP2 AGC DC voltage measures 71 mV, instead of > the expected >4.5 V. I also checked the 60 MHz signal into 50 ohms and > see about 30 mW. One place in the manual says this should be >1V (which > it is), but another place says 300 mW. Is loading this with 50 ohms > correct? If so, maybe 300 mW is a typo and it should read 30 mW? In any > event the AGC DC voltage is wrong and I suppose needs to be addressed. > > The second issue, and probably a more serious one, is that the PHOTO I > meter reading is 20 (less than the specified 25) and the actual current > measured out of A12 is -10 uA (should be 50 uA or greater). There are > two issues here: (1) the actual current disagrees with the meter > indication and (2) the current is low. Is the low current possibly > explained by cell flooding and or incorrect/insufficient 60 MHz RF > signal application? > > Thoughts on next steps? I believe I need to address both issues > above. Addressing cell flooding is straightforward, but will take some > time. The troubleshooting section for A3 says to perform the loop > adjustment procedure if the AGC DC voltage is wrong. Which should I do > first? Or, should I start with something else? > > Thanks! > Matt
MH
Matt Huszagh
Sun, Nov 5, 2023 7:29 PM

I've decide to start by addressing the cell flooding. Despite the
incorrect AGC DC voltage, the 60 MHz signal looks roughly ok and from a
quick look I don't think I can really perform the loop alignment without
some sort of RVFR signal.

I've attached a current-limited supply at 5V through a 5ohm power
resistor to the TED wires and have the unit powered with the meter set
to 2nd harmonic. The meter indication almost immediately started
deflecting off 0. It's been just a few minutes and the readings are
getting almost to 10 at max deflection, though deflecting back and
forth. I'm hopeful this is a good sign. I'll update as this unit has
more time to de-flood.

Please feel free to make suggestions if I should be following a
different approach at the moment.

Thanks!
Matt

Matt Huszagh huszaghmatt@gmail.com writes:

There are a few other things I wanted to mention.

I checked the DC voltages around the cell and lamp heaters. J15P3 (cell
heater) measured 15.4V and J15P5 measured 14.4V. P4 measured 28.6V. The
specified nominal voltages for P3 and P5 are 18V, so these are a bit
off, but since the heater currents look ok as measured by the front
panel meter, I got the sense that this is probably ok. Thoughts?

Also, as part of the pre-emptive maintenance I replaced the lamp
assembly film resistor and, in the process, disassembled part of A12
including the lamp assembly. Everything looked fine to my eye and I have
pictures I can share if desired. In particular, the lamp assembly was
not charred.

Matt

Matt Huszagh huszaghmatt@gmail.com writes:

Hi,

I have an HP 5065A received in unknown condition. After performing some
pre-emptive maintenance (basically Corby's recommendations here:
https://www.eevblog.com/forum/metrology/inspecting-aligning-and-testing-an-hp-5065a-rubidium/
and the first aid advice from Luciano here:
http://www.timeok.it/hp5065a-corner-3/) I turned the unit on. +/- 20V
supplies look good and cell and lamp ovens drop down from pegged in
under 45 min to about 32 each.

However, the 2nd harmonic reading is 0 and of course the unit does not
lock.

This unit was left unpowered for some time (I got the sense it was at
least several years) before I received it, and therefore cell flooding
is likely. However, I have noticed a few other discrepencies.

The first is that the A3 TP2 AGC DC voltage measures 71 mV, instead of
the expected >4.5 V. I also checked the 60 MHz signal into 50 ohms and
see about 30 mW. One place in the manual says this should be >1V (which
it is), but another place says 300 mW. Is loading this with 50 ohms
correct? If so, maybe 300 mW is a typo and it should read 30 mW? In any
event the AGC DC voltage is wrong and I suppose needs to be addressed.

The second issue, and probably a more serious one, is that the PHOTO I
meter reading is 20 (less than the specified 25) and the actual current
measured out of A12 is -10 uA (should be 50 uA or greater). There are
two issues here: (1) the actual current disagrees with the meter
indication and (2) the current is low. Is the low current possibly
explained by cell flooding and or incorrect/insufficient 60 MHz RF
signal application?

Thoughts on next steps? I believe I need to address both issues
above. Addressing cell flooding is straightforward, but will take some
time. The troubleshooting section for A3 says to perform the loop
adjustment procedure if the AGC DC voltage is wrong. Which should I do
first? Or, should I start with something else?

Thanks!
Matt

I've decide to start by addressing the cell flooding. Despite the incorrect AGC DC voltage, the 60 MHz signal looks roughly ok and from a quick look I don't think I can really perform the loop alignment without some sort of RVFR signal. I've attached a current-limited supply at 5V through a 5ohm power resistor to the TED wires and have the unit powered with the meter set to 2nd harmonic. The meter indication almost immediately started deflecting off 0. It's been just a few minutes and the readings are getting almost to 10 at max deflection, though deflecting back and forth. I'm hopeful this is a good sign. I'll update as this unit has more time to de-flood. Please feel free to make suggestions if I should be following a different approach at the moment. Thanks! Matt Matt Huszagh <huszaghmatt@gmail.com> writes: > There are a few other things I wanted to mention. > > I checked the DC voltages around the cell and lamp heaters. J15P3 (cell > heater) measured 15.4V and J15P5 measured 14.4V. P4 measured 28.6V. The > specified nominal voltages for P3 and P5 are 18V, so these are a bit > off, but since the heater currents look ok as measured by the front > panel meter, I got the sense that this is probably ok. Thoughts? > > Also, as part of the pre-emptive maintenance I replaced the lamp > assembly film resistor and, in the process, disassembled part of A12 > including the lamp assembly. Everything looked fine to my eye and I have > pictures I can share if desired. In particular, the lamp assembly was > not charred. > > Matt > > Matt Huszagh <huszaghmatt@gmail.com> writes: > >> Hi, >> >> I have an HP 5065A received in unknown condition. After performing some >> pre-emptive maintenance (basically Corby's recommendations here: >> https://www.eevblog.com/forum/metrology/inspecting-aligning-and-testing-an-hp-5065a-rubidium/ >> and the first aid advice from Luciano here: >> http://www.timeok.it/hp5065a-corner-3/) I turned the unit on. +/- 20V >> supplies look good and cell and lamp ovens drop down from pegged in >> under 45 min to about 32 each. >> >> However, the 2nd harmonic reading is 0 and of course the unit does not >> lock. >> >> This unit was left unpowered for some time (I got the sense it was at >> least several years) before I received it, and therefore cell flooding >> is likely. However, I have noticed a few other discrepencies. >> >> The first is that the A3 TP2 AGC DC voltage measures 71 mV, instead of >> the expected >4.5 V. I also checked the 60 MHz signal into 50 ohms and >> see about 30 mW. One place in the manual says this should be >1V (which >> it is), but another place says 300 mW. Is loading this with 50 ohms >> correct? If so, maybe 300 mW is a typo and it should read 30 mW? In any >> event the AGC DC voltage is wrong and I suppose needs to be addressed. >> >> The second issue, and probably a more serious one, is that the PHOTO I >> meter reading is 20 (less than the specified 25) and the actual current >> measured out of A12 is -10 uA (should be 50 uA or greater). There are >> two issues here: (1) the actual current disagrees with the meter >> indication and (2) the current is low. Is the low current possibly >> explained by cell flooding and or incorrect/insufficient 60 MHz RF >> signal application? >> >> Thoughts on next steps? I believe I need to address both issues >> above. Addressing cell flooding is straightforward, but will take some >> time. The troubleshooting section for A3 says to perform the loop >> adjustment procedure if the AGC DC voltage is wrong. Which should I do >> first? Or, should I start with something else? >> >> Thanks! >> Matt
BC
Bob Camp
Mon, Nov 6, 2023 1:46 PM

Hi

If you haven’t already checked the 100 uf “wet” tantilum caps, it would be a really good idea to do so. It’s an eyeball sort of thing. If there is a pile of greenish rust instead of one cap lead …. time to replace it.

Bob

On Nov 5, 2023, at 2:29 PM, Matt Huszagh via time-nuts time-nuts@lists.febo.com wrote:

I've decide to start by addressing the cell flooding. Despite the
incorrect AGC DC voltage, the 60 MHz signal looks roughly ok and from a
quick look I don't think I can really perform the loop alignment without
some sort of RVFR signal.

I've attached a current-limited supply at 5V through a 5ohm power
resistor to the TED wires and have the unit powered with the meter set
to 2nd harmonic. The meter indication almost immediately started
deflecting off 0. It's been just a few minutes and the readings are
getting almost to 10 at max deflection, though deflecting back and
forth. I'm hopeful this is a good sign. I'll update as this unit has
more time to de-flood.

Please feel free to make suggestions if I should be following a
different approach at the moment.

Thanks!
Matt

Matt Huszagh huszaghmatt@gmail.com writes:

There are a few other things I wanted to mention.

I checked the DC voltages around the cell and lamp heaters. J15P3 (cell
heater) measured 15.4V and J15P5 measured 14.4V. P4 measured 28.6V. The
specified nominal voltages for P3 and P5 are 18V, so these are a bit
off, but since the heater currents look ok as measured by the front
panel meter, I got the sense that this is probably ok. Thoughts?

Also, as part of the pre-emptive maintenance I replaced the lamp
assembly film resistor and, in the process, disassembled part of A12
including the lamp assembly. Everything looked fine to my eye and I have
pictures I can share if desired. In particular, the lamp assembly was
not charred.

Matt

Matt Huszagh huszaghmatt@gmail.com writes:

Hi,

I have an HP 5065A received in unknown condition. After performing some
pre-emptive maintenance (basically Corby's recommendations here:
https://www.eevblog.com/forum/metrology/inspecting-aligning-and-testing-an-hp-5065a-rubidium/
and the first aid advice from Luciano here:
http://www.timeok.it/hp5065a-corner-3/) I turned the unit on. +/- 20V
supplies look good and cell and lamp ovens drop down from pegged in
under 45 min to about 32 each.

However, the 2nd harmonic reading is 0 and of course the unit does not
lock.

This unit was left unpowered for some time (I got the sense it was at
least several years) before I received it, and therefore cell flooding
is likely. However, I have noticed a few other discrepencies.

The first is that the A3 TP2 AGC DC voltage measures 71 mV, instead of
the expected >4.5 V. I also checked the 60 MHz signal into 50 ohms and
see about 30 mW. One place in the manual says this should be >1V (which
it is), but another place says 300 mW. Is loading this with 50 ohms
correct? If so, maybe 300 mW is a typo and it should read 30 mW? In any
event the AGC DC voltage is wrong and I suppose needs to be addressed.

The second issue, and probably a more serious one, is that the PHOTO I
meter reading is 20 (less than the specified 25) and the actual current
measured out of A12 is -10 uA (should be 50 uA or greater). There are
two issues here: (1) the actual current disagrees with the meter
indication and (2) the current is low. Is the low current possibly
explained by cell flooding and or incorrect/insufficient 60 MHz RF
signal application?

Thoughts on next steps? I believe I need to address both issues
above. Addressing cell flooding is straightforward, but will take some
time. The troubleshooting section for A3 says to perform the loop
adjustment procedure if the AGC DC voltage is wrong. Which should I do
first? Or, should I start with something else?

Thanks!
Matt


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

Hi If you haven’t already checked the 100 uf “wet” tantilum caps, it would be a really good idea to do so. It’s an eyeball sort of thing. If there is a pile of greenish rust instead of one cap lead …. time to replace it. Bob > On Nov 5, 2023, at 2:29 PM, Matt Huszagh via time-nuts <time-nuts@lists.febo.com> wrote: > > I've decide to start by addressing the cell flooding. Despite the > incorrect AGC DC voltage, the 60 MHz signal looks roughly ok and from a > quick look I don't think I can really perform the loop alignment without > some sort of RVFR signal. > > I've attached a current-limited supply at 5V through a 5ohm power > resistor to the TED wires and have the unit powered with the meter set > to 2nd harmonic. The meter indication almost immediately started > deflecting off 0. It's been just a few minutes and the readings are > getting almost to 10 at max deflection, though deflecting back and > forth. I'm hopeful this is a good sign. I'll update as this unit has > more time to de-flood. > > Please feel free to make suggestions if I should be following a > different approach at the moment. > > Thanks! > Matt > > Matt Huszagh <huszaghmatt@gmail.com> writes: > >> There are a few other things I wanted to mention. >> >> I checked the DC voltages around the cell and lamp heaters. J15P3 (cell >> heater) measured 15.4V and J15P5 measured 14.4V. P4 measured 28.6V. The >> specified nominal voltages for P3 and P5 are 18V, so these are a bit >> off, but since the heater currents look ok as measured by the front >> panel meter, I got the sense that this is probably ok. Thoughts? >> >> Also, as part of the pre-emptive maintenance I replaced the lamp >> assembly film resistor and, in the process, disassembled part of A12 >> including the lamp assembly. Everything looked fine to my eye and I have >> pictures I can share if desired. In particular, the lamp assembly was >> not charred. >> >> Matt >> >> Matt Huszagh <huszaghmatt@gmail.com> writes: >> >>> Hi, >>> >>> I have an HP 5065A received in unknown condition. After performing some >>> pre-emptive maintenance (basically Corby's recommendations here: >>> https://www.eevblog.com/forum/metrology/inspecting-aligning-and-testing-an-hp-5065a-rubidium/ >>> and the first aid advice from Luciano here: >>> http://www.timeok.it/hp5065a-corner-3/) I turned the unit on. +/- 20V >>> supplies look good and cell and lamp ovens drop down from pegged in >>> under 45 min to about 32 each. >>> >>> However, the 2nd harmonic reading is 0 and of course the unit does not >>> lock. >>> >>> This unit was left unpowered for some time (I got the sense it was at >>> least several years) before I received it, and therefore cell flooding >>> is likely. However, I have noticed a few other discrepencies. >>> >>> The first is that the A3 TP2 AGC DC voltage measures 71 mV, instead of >>> the expected >4.5 V. I also checked the 60 MHz signal into 50 ohms and >>> see about 30 mW. One place in the manual says this should be >1V (which >>> it is), but another place says 300 mW. Is loading this with 50 ohms >>> correct? If so, maybe 300 mW is a typo and it should read 30 mW? In any >>> event the AGC DC voltage is wrong and I suppose needs to be addressed. >>> >>> The second issue, and probably a more serious one, is that the PHOTO I >>> meter reading is 20 (less than the specified 25) and the actual current >>> measured out of A12 is -10 uA (should be 50 uA or greater). There are >>> two issues here: (1) the actual current disagrees with the meter >>> indication and (2) the current is low. Is the low current possibly >>> explained by cell flooding and or incorrect/insufficient 60 MHz RF >>> signal application? >>> >>> Thoughts on next steps? I believe I need to address both issues >>> above. Addressing cell flooding is straightforward, but will take some >>> time. The troubleshooting section for A3 says to perform the loop >>> adjustment procedure if the AGC DC voltage is wrong. Which should I do >>> first? Or, should I start with something else? >>> >>> Thanks! >>> Matt > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com
MH
Matt Huszagh
Mon, Nov 6, 2023 5:55 PM

Update: I realized my unit had the shunt inductor across the TED wires
so I cut this out and am now back at the de-flooding process.

When I first started the de-flooding process yesterday, I noticed quite
a bit of 2nd harmonic meter deflection/activity, but now it's pretty
steady at 0. I'm not sure if this is normal. But, I'm not very far into
the 6-day process, so I suppose I just need to wait.

Matt

Matt Huszagh huszaghmatt@gmail.com writes:

I've decide to start by addressing the cell flooding. Despite the
incorrect AGC DC voltage, the 60 MHz signal looks roughly ok and from a
quick look I don't think I can really perform the loop alignment without
some sort of RVFR signal.

I've attached a current-limited supply at 5V through a 5ohm power
resistor to the TED wires and have the unit powered with the meter set
to 2nd harmonic. The meter indication almost immediately started
deflecting off 0. It's been just a few minutes and the readings are
getting almost to 10 at max deflection, though deflecting back and
forth. I'm hopeful this is a good sign. I'll update as this unit has
more time to de-flood.

Please feel free to make suggestions if I should be following a
different approach at the moment.

Thanks!
Matt

Matt Huszagh huszaghmatt@gmail.com writes:

There are a few other things I wanted to mention.

I checked the DC voltages around the cell and lamp heaters. J15P3 (cell
heater) measured 15.4V and J15P5 measured 14.4V. P4 measured 28.6V. The
specified nominal voltages for P3 and P5 are 18V, so these are a bit
off, but since the heater currents look ok as measured by the front
panel meter, I got the sense that this is probably ok. Thoughts?

Also, as part of the pre-emptive maintenance I replaced the lamp
assembly film resistor and, in the process, disassembled part of A12
including the lamp assembly. Everything looked fine to my eye and I have
pictures I can share if desired. In particular, the lamp assembly was
not charred.

Matt

Matt Huszagh huszaghmatt@gmail.com writes:

Hi,

I have an HP 5065A received in unknown condition. After performing some
pre-emptive maintenance (basically Corby's recommendations here:
https://www.eevblog.com/forum/metrology/inspecting-aligning-and-testing-an-hp-5065a-rubidium/
and the first aid advice from Luciano here:
http://www.timeok.it/hp5065a-corner-3/) I turned the unit on. +/- 20V
supplies look good and cell and lamp ovens drop down from pegged in
under 45 min to about 32 each.

However, the 2nd harmonic reading is 0 and of course the unit does not
lock.

This unit was left unpowered for some time (I got the sense it was at
least several years) before I received it, and therefore cell flooding
is likely. However, I have noticed a few other discrepencies.

The first is that the A3 TP2 AGC DC voltage measures 71 mV, instead of
the expected >4.5 V. I also checked the 60 MHz signal into 50 ohms and
see about 30 mW. One place in the manual says this should be >1V (which
it is), but another place says 300 mW. Is loading this with 50 ohms
correct? If so, maybe 300 mW is a typo and it should read 30 mW? In any
event the AGC DC voltage is wrong and I suppose needs to be addressed.

The second issue, and probably a more serious one, is that the PHOTO I
meter reading is 20 (less than the specified 25) and the actual current
measured out of A12 is -10 uA (should be 50 uA or greater). There are
two issues here: (1) the actual current disagrees with the meter
indication and (2) the current is low. Is the low current possibly
explained by cell flooding and or incorrect/insufficient 60 MHz RF
signal application?

Thoughts on next steps? I believe I need to address both issues
above. Addressing cell flooding is straightforward, but will take some
time. The troubleshooting section for A3 says to perform the loop
adjustment procedure if the AGC DC voltage is wrong. Which should I do
first? Or, should I start with something else?

Thanks!
Matt

Update: I realized my unit had the shunt inductor across the TED wires so I cut this out and am now back at the de-flooding process. When I first started the de-flooding process yesterday, I noticed quite a bit of 2nd harmonic meter deflection/activity, but now it's pretty steady at 0. I'm not sure if this is normal. But, I'm not very far into the 6-day process, so I suppose I just need to wait. Matt Matt Huszagh <huszaghmatt@gmail.com> writes: > I've decide to start by addressing the cell flooding. Despite the > incorrect AGC DC voltage, the 60 MHz signal looks roughly ok and from a > quick look I don't think I can really perform the loop alignment without > some sort of RVFR signal. > > I've attached a current-limited supply at 5V through a 5ohm power > resistor to the TED wires and have the unit powered with the meter set > to 2nd harmonic. The meter indication almost immediately started > deflecting off 0. It's been just a few minutes and the readings are > getting almost to 10 at max deflection, though deflecting back and > forth. I'm hopeful this is a good sign. I'll update as this unit has > more time to de-flood. > > Please feel free to make suggestions if I should be following a > different approach at the moment. > > Thanks! > Matt > > Matt Huszagh <huszaghmatt@gmail.com> writes: > >> There are a few other things I wanted to mention. >> >> I checked the DC voltages around the cell and lamp heaters. J15P3 (cell >> heater) measured 15.4V and J15P5 measured 14.4V. P4 measured 28.6V. The >> specified nominal voltages for P3 and P5 are 18V, so these are a bit >> off, but since the heater currents look ok as measured by the front >> panel meter, I got the sense that this is probably ok. Thoughts? >> >> Also, as part of the pre-emptive maintenance I replaced the lamp >> assembly film resistor and, in the process, disassembled part of A12 >> including the lamp assembly. Everything looked fine to my eye and I have >> pictures I can share if desired. In particular, the lamp assembly was >> not charred. >> >> Matt >> >> Matt Huszagh <huszaghmatt@gmail.com> writes: >> >>> Hi, >>> >>> I have an HP 5065A received in unknown condition. After performing some >>> pre-emptive maintenance (basically Corby's recommendations here: >>> https://www.eevblog.com/forum/metrology/inspecting-aligning-and-testing-an-hp-5065a-rubidium/ >>> and the first aid advice from Luciano here: >>> http://www.timeok.it/hp5065a-corner-3/) I turned the unit on. +/- 20V >>> supplies look good and cell and lamp ovens drop down from pegged in >>> under 45 min to about 32 each. >>> >>> However, the 2nd harmonic reading is 0 and of course the unit does not >>> lock. >>> >>> This unit was left unpowered for some time (I got the sense it was at >>> least several years) before I received it, and therefore cell flooding >>> is likely. However, I have noticed a few other discrepencies. >>> >>> The first is that the A3 TP2 AGC DC voltage measures 71 mV, instead of >>> the expected >4.5 V. I also checked the 60 MHz signal into 50 ohms and >>> see about 30 mW. One place in the manual says this should be >1V (which >>> it is), but another place says 300 mW. Is loading this with 50 ohms >>> correct? If so, maybe 300 mW is a typo and it should read 30 mW? In any >>> event the AGC DC voltage is wrong and I suppose needs to be addressed. >>> >>> The second issue, and probably a more serious one, is that the PHOTO I >>> meter reading is 20 (less than the specified 25) and the actual current >>> measured out of A12 is -10 uA (should be 50 uA or greater). There are >>> two issues here: (1) the actual current disagrees with the meter >>> indication and (2) the current is low. Is the low current possibly >>> explained by cell flooding and or incorrect/insufficient 60 MHz RF >>> signal application? >>> >>> Thoughts on next steps? I believe I need to address both issues >>> above. Addressing cell flooding is straightforward, but will take some >>> time. The troubleshooting section for A3 says to perform the loop >>> adjustment procedure if the AGC DC voltage is wrong. Which should I do >>> first? Or, should I start with something else? >>> >>> Thanks! >>> Matt
MH
Matt Huszagh
Tue, Nov 7, 2023 10:42 PM

I'm back to troubleshooting the A3 multiplier assembly because if the
60 + 5.315... MHz signal to the SRD isn't correct, I wouldn't expect to
know if and when de-flooding succeeds.

The 5.315... MHz component is a bit weak - in the neighborhood of 5mVrms
instead of the expected 20mVrms. Additionally, the AGC DC voltage (TP2)
is 70 mV, but should be around 4.5 V.

I've got A3 hooked up in isolation on my bench with 5MHz at 1Vrms to J2
and 600mVrms to J5 at 5.315MHz. Any suggestions for how to troubleshoot
this? I measured Q4 and C12 to ensure they're not shorted - they're
fine. I also checked different points in the circuit and compared them
with the voltage annotations in the schematic. Unfortunately, I'm having
some trouble understanding this circuit, and some of the annotations
confuse me.

According to the annotations, the base of Q4 (the AGC amplifier), is at
+0.7Vdc for no AGC and +6Vdc for full AGC. But there's no emitter
resistor so how does this get to +6Vdc? Moreover, I don't see a DC path
to supply current to raise the DC voltage. In case it's helpful, Q4 is
apparently a 2N2484.

Another example is Q9 base is labelled with 2.4V. I think that's dc
since I see Vrms in a number of other places. But I don't see how that
could be true. This is fed via a +20V supply, and the base is biased by
a resistive dividier with 18k and 2k, so it should be at most +2V, but
then there's the parallel path of the BE junction and emitter resistor,
so it would be even lower than this. So, where are they getting 2.4V?

I checked a number of points in the main signal path and some looked
more or less in line with amplitude annotations, but others didn't. I'm
not totally sure how precise these annotations are supposed to be,
though.

Finally, I wanted to clarify how the RVFR SRD bias works. The only DC
path to the bias test point is through J4 (which provides the
60 + 5.315... MHz signal to the SRD). But, the RVFR schematic does not
show any DC bias applied to the SRD. I presume that's there but just not
shown. Is that right? That would make sense, since the pot there (R40)
would divide down the bias.

Any suggestions would be immensely appreciated.

Thank you!
Matt

Matt Huszagh huszaghmatt@gmail.com writes:

I've decide to start by addressing the cell flooding. Despite the
incorrect AGC DC voltage, the 60 MHz signal looks roughly ok and from a
quick look I don't think I can really perform the loop alignment without
some sort of RVFR signal.

I've attached a current-limited supply at 5V through a 5ohm power
resistor to the TED wires and have the unit powered with the meter set
to 2nd harmonic. The meter indication almost immediately started
deflecting off 0. It's been just a few minutes and the readings are
getting almost to 10 at max deflection, though deflecting back and
forth. I'm hopeful this is a good sign. I'll update as this unit has
more time to de-flood.

Please feel free to make suggestions if I should be following a
different approach at the moment.

Thanks!
Matt

Matt Huszagh huszaghmatt@gmail.com writes:

There are a few other things I wanted to mention.

I checked the DC voltages around the cell and lamp heaters. J15P3 (cell
heater) measured 15.4V and J15P5 measured 14.4V. P4 measured 28.6V. The
specified nominal voltages for P3 and P5 are 18V, so these are a bit
off, but since the heater currents look ok as measured by the front
panel meter, I got the sense that this is probably ok. Thoughts?

Also, as part of the pre-emptive maintenance I replaced the lamp
assembly film resistor and, in the process, disassembled part of A12
including the lamp assembly. Everything looked fine to my eye and I have
pictures I can share if desired. In particular, the lamp assembly was
not charred.

Matt

Matt Huszagh huszaghmatt@gmail.com writes:

Hi,

I have an HP 5065A received in unknown condition. After performing some
pre-emptive maintenance (basically Corby's recommendations here:
https://www.eevblog.com/forum/metrology/inspecting-aligning-and-testing-an-hp-5065a-rubidium/
and the first aid advice from Luciano here:
http://www.timeok.it/hp5065a-corner-3/) I turned the unit on. +/- 20V
supplies look good and cell and lamp ovens drop down from pegged in
under 45 min to about 32 each.

However, the 2nd harmonic reading is 0 and of course the unit does not
lock.

This unit was left unpowered for some time (I got the sense it was at
least several years) before I received it, and therefore cell flooding
is likely. However, I have noticed a few other discrepencies.

The first is that the A3 TP2 AGC DC voltage measures 71 mV, instead of
the expected >4.5 V. I also checked the 60 MHz signal into 50 ohms and
see about 30 mW. One place in the manual says this should be >1V (which
it is), but another place says 300 mW. Is loading this with 50 ohms
correct? If so, maybe 300 mW is a typo and it should read 30 mW? In any
event the AGC DC voltage is wrong and I suppose needs to be addressed.

The second issue, and probably a more serious one, is that the PHOTO I
meter reading is 20 (less than the specified 25) and the actual current
measured out of A12 is -10 uA (should be 50 uA or greater). There are
two issues here: (1) the actual current disagrees with the meter
indication and (2) the current is low. Is the low current possibly
explained by cell flooding and or incorrect/insufficient 60 MHz RF
signal application?

Thoughts on next steps? I believe I need to address both issues
above. Addressing cell flooding is straightforward, but will take some
time. The troubleshooting section for A3 says to perform the loop
adjustment procedure if the AGC DC voltage is wrong. Which should I do
first? Or, should I start with something else?

Thanks!
Matt

I'm back to troubleshooting the A3 multiplier assembly because if the 60 + 5.315... MHz signal to the SRD isn't correct, I wouldn't expect to know if and when de-flooding succeeds. The 5.315... MHz component is a bit weak - in the neighborhood of 5mVrms instead of the expected 20mVrms. Additionally, the AGC DC voltage (TP2) is 70 mV, but should be around 4.5 V. I've got A3 hooked up in isolation on my bench with 5MHz at 1Vrms to J2 and 600mVrms to J5 at 5.315MHz. Any suggestions for how to troubleshoot this? I measured Q4 and C12 to ensure they're not shorted - they're fine. I also checked different points in the circuit and compared them with the voltage annotations in the schematic. Unfortunately, I'm having some trouble understanding this circuit, and some of the annotations confuse me. According to the annotations, the base of Q4 (the AGC amplifier), is at +0.7Vdc for no AGC and +6Vdc for full AGC. But there's no emitter resistor so how does this get to +6Vdc? Moreover, I don't see a DC path to supply current to raise the DC voltage. In case it's helpful, Q4 is apparently a 2N2484. Another example is Q9 base is labelled with 2.4V. I think that's dc since I see Vrms in a number of other places. But I don't see how that could be true. This is fed via a +20V supply, and the base is biased by a resistive dividier with 18k and 2k, so it should be at most +2V, but then there's the parallel path of the BE junction and emitter resistor, so it would be even lower than this. So, where are they getting 2.4V? I checked a number of points in the main signal path and some looked more or less in line with amplitude annotations, but others didn't. I'm not totally sure how precise these annotations are supposed to be, though. Finally, I wanted to clarify how the RVFR SRD bias works. The only DC path to the bias test point is through J4 (which provides the 60 + 5.315... MHz signal to the SRD). But, the RVFR schematic does not show any DC bias applied to the SRD. I presume that's there but just not shown. Is that right? That would make sense, since the pot there (R40) would divide down the bias. Any suggestions would be immensely appreciated. Thank you! Matt Matt Huszagh <huszaghmatt@gmail.com> writes: > I've decide to start by addressing the cell flooding. Despite the > incorrect AGC DC voltage, the 60 MHz signal looks roughly ok and from a > quick look I don't think I can really perform the loop alignment without > some sort of RVFR signal. > > I've attached a current-limited supply at 5V through a 5ohm power > resistor to the TED wires and have the unit powered with the meter set > to 2nd harmonic. The meter indication almost immediately started > deflecting off 0. It's been just a few minutes and the readings are > getting almost to 10 at max deflection, though deflecting back and > forth. I'm hopeful this is a good sign. I'll update as this unit has > more time to de-flood. > > Please feel free to make suggestions if I should be following a > different approach at the moment. > > Thanks! > Matt > > Matt Huszagh <huszaghmatt@gmail.com> writes: > >> There are a few other things I wanted to mention. >> >> I checked the DC voltages around the cell and lamp heaters. J15P3 (cell >> heater) measured 15.4V and J15P5 measured 14.4V. P4 measured 28.6V. The >> specified nominal voltages for P3 and P5 are 18V, so these are a bit >> off, but since the heater currents look ok as measured by the front >> panel meter, I got the sense that this is probably ok. Thoughts? >> >> Also, as part of the pre-emptive maintenance I replaced the lamp >> assembly film resistor and, in the process, disassembled part of A12 >> including the lamp assembly. Everything looked fine to my eye and I have >> pictures I can share if desired. In particular, the lamp assembly was >> not charred. >> >> Matt >> >> Matt Huszagh <huszaghmatt@gmail.com> writes: >> >>> Hi, >>> >>> I have an HP 5065A received in unknown condition. After performing some >>> pre-emptive maintenance (basically Corby's recommendations here: >>> https://www.eevblog.com/forum/metrology/inspecting-aligning-and-testing-an-hp-5065a-rubidium/ >>> and the first aid advice from Luciano here: >>> http://www.timeok.it/hp5065a-corner-3/) I turned the unit on. +/- 20V >>> supplies look good and cell and lamp ovens drop down from pegged in >>> under 45 min to about 32 each. >>> >>> However, the 2nd harmonic reading is 0 and of course the unit does not >>> lock. >>> >>> This unit was left unpowered for some time (I got the sense it was at >>> least several years) before I received it, and therefore cell flooding >>> is likely. However, I have noticed a few other discrepencies. >>> >>> The first is that the A3 TP2 AGC DC voltage measures 71 mV, instead of >>> the expected >4.5 V. I also checked the 60 MHz signal into 50 ohms and >>> see about 30 mW. One place in the manual says this should be >1V (which >>> it is), but another place says 300 mW. Is loading this with 50 ohms >>> correct? If so, maybe 300 mW is a typo and it should read 30 mW? In any >>> event the AGC DC voltage is wrong and I suppose needs to be addressed. >>> >>> The second issue, and probably a more serious one, is that the PHOTO I >>> meter reading is 20 (less than the specified 25) and the actual current >>> measured out of A12 is -10 uA (should be 50 uA or greater). There are >>> two issues here: (1) the actual current disagrees with the meter >>> indication and (2) the current is low. Is the low current possibly >>> explained by cell flooding and or incorrect/insufficient 60 MHz RF >>> signal application? >>> >>> Thoughts on next steps? I believe I need to address both issues >>> above. Addressing cell flooding is straightforward, but will take some >>> time. The troubleshooting section for A3 says to perform the loop >>> adjustment procedure if the AGC DC voltage is wrong. Which should I do >>> first? Or, should I start with something else? >>> >>> Thanks! >>> Matt
MH
Matt Huszagh
Wed, Nov 8, 2023 3:52 AM

Well I think I was wrong about A3 being an issue. I compared this unit
to a working unit and they behave pretty similarly. The A1 synthesizer
5.135 MHz output is a little lower on the (thus far) non-functioning
unit, but it's not a huge difference and the RF excitation signal into
the RVFR looks pretty similar. A3TP2 is low on both units (93 mV on the
working one). I expect the working unit probably just needs an
alignment.

Anyway, I'm back to de-flooding.

Matt

Matt Huszagh huszaghmatt@gmail.com writes:

I'm back to troubleshooting the A3 multiplier assembly because if the
60 + 5.315... MHz signal to the SRD isn't correct, I wouldn't expect to
know if and when de-flooding succeeds.

The 5.315... MHz component is a bit weak - in the neighborhood of 5mVrms
instead of the expected 20mVrms. Additionally, the AGC DC voltage (TP2)
is 70 mV, but should be around 4.5 V.

I've got A3 hooked up in isolation on my bench with 5MHz at 1Vrms to J2
and 600mVrms to J5 at 5.315MHz. Any suggestions for how to troubleshoot
this? I measured Q4 and C12 to ensure they're not shorted - they're
fine. I also checked different points in the circuit and compared them
with the voltage annotations in the schematic. Unfortunately, I'm having
some trouble understanding this circuit, and some of the annotations
confuse me.

According to the annotations, the base of Q4 (the AGC amplifier), is at
+0.7Vdc for no AGC and +6Vdc for full AGC. But there's no emitter
resistor so how does this get to +6Vdc? Moreover, I don't see a DC path
to supply current to raise the DC voltage. In case it's helpful, Q4 is
apparently a 2N2484.

Another example is Q9 base is labelled with 2.4V. I think that's dc
since I see Vrms in a number of other places. But I don't see how that
could be true. This is fed via a +20V supply, and the base is biased by
a resistive dividier with 18k and 2k, so it should be at most +2V, but
then there's the parallel path of the BE junction and emitter resistor,
so it would be even lower than this. So, where are they getting 2.4V?

I checked a number of points in the main signal path and some looked
more or less in line with amplitude annotations, but others didn't. I'm
not totally sure how precise these annotations are supposed to be,
though.

Finally, I wanted to clarify how the RVFR SRD bias works. The only DC
path to the bias test point is through J4 (which provides the
60 + 5.315... MHz signal to the SRD). But, the RVFR schematic does not
show any DC bias applied to the SRD. I presume that's there but just not
shown. Is that right? That would make sense, since the pot there (R40)
would divide down the bias.

Any suggestions would be immensely appreciated.

Thank you!
Matt

Matt Huszagh huszaghmatt@gmail.com writes:

I've decide to start by addressing the cell flooding. Despite the
incorrect AGC DC voltage, the 60 MHz signal looks roughly ok and from a
quick look I don't think I can really perform the loop alignment without
some sort of RVFR signal.

I've attached a current-limited supply at 5V through a 5ohm power
resistor to the TED wires and have the unit powered with the meter set
to 2nd harmonic. The meter indication almost immediately started
deflecting off 0. It's been just a few minutes and the readings are
getting almost to 10 at max deflection, though deflecting back and
forth. I'm hopeful this is a good sign. I'll update as this unit has
more time to de-flood.

Please feel free to make suggestions if I should be following a
different approach at the moment.

Thanks!
Matt

Matt Huszagh huszaghmatt@gmail.com writes:

There are a few other things I wanted to mention.

I checked the DC voltages around the cell and lamp heaters. J15P3 (cell
heater) measured 15.4V and J15P5 measured 14.4V. P4 measured 28.6V. The
specified nominal voltages for P3 and P5 are 18V, so these are a bit
off, but since the heater currents look ok as measured by the front
panel meter, I got the sense that this is probably ok. Thoughts?

Also, as part of the pre-emptive maintenance I replaced the lamp
assembly film resistor and, in the process, disassembled part of A12
including the lamp assembly. Everything looked fine to my eye and I have
pictures I can share if desired. In particular, the lamp assembly was
not charred.

Matt

Matt Huszagh huszaghmatt@gmail.com writes:

Hi,

I have an HP 5065A received in unknown condition. After performing some
pre-emptive maintenance (basically Corby's recommendations here:
https://www.eevblog.com/forum/metrology/inspecting-aligning-and-testing-an-hp-5065a-rubidium/
and the first aid advice from Luciano here:
http://www.timeok.it/hp5065a-corner-3/) I turned the unit on. +/- 20V
supplies look good and cell and lamp ovens drop down from pegged in
under 45 min to about 32 each.

However, the 2nd harmonic reading is 0 and of course the unit does not
lock.

This unit was left unpowered for some time (I got the sense it was at
least several years) before I received it, and therefore cell flooding
is likely. However, I have noticed a few other discrepencies.

The first is that the A3 TP2 AGC DC voltage measures 71 mV, instead of
the expected >4.5 V. I also checked the 60 MHz signal into 50 ohms and
see about 30 mW. One place in the manual says this should be >1V (which
it is), but another place says 300 mW. Is loading this with 50 ohms
correct? If so, maybe 300 mW is a typo and it should read 30 mW? In any
event the AGC DC voltage is wrong and I suppose needs to be addressed.

The second issue, and probably a more serious one, is that the PHOTO I
meter reading is 20 (less than the specified 25) and the actual current
measured out of A12 is -10 uA (should be 50 uA or greater). There are
two issues here: (1) the actual current disagrees with the meter
indication and (2) the current is low. Is the low current possibly
explained by cell flooding and or incorrect/insufficient 60 MHz RF
signal application?

Thoughts on next steps? I believe I need to address both issues
above. Addressing cell flooding is straightforward, but will take some
time. The troubleshooting section for A3 says to perform the loop
adjustment procedure if the AGC DC voltage is wrong. Which should I do
first? Or, should I start with something else?

Thanks!
Matt

Well I think I was wrong about A3 being an issue. I compared this unit to a working unit and they behave pretty similarly. The A1 synthesizer 5.135 MHz output is a little lower on the (thus far) non-functioning unit, but it's not a huge difference and the RF excitation signal into the RVFR looks pretty similar. A3TP2 is low on both units (93 mV on the working one). I expect the working unit probably just needs an alignment. Anyway, I'm back to de-flooding. Matt Matt Huszagh <huszaghmatt@gmail.com> writes: > I'm back to troubleshooting the A3 multiplier assembly because if the > 60 + 5.315... MHz signal to the SRD isn't correct, I wouldn't expect to > know if and when de-flooding succeeds. > > The 5.315... MHz component is a bit weak - in the neighborhood of 5mVrms > instead of the expected 20mVrms. Additionally, the AGC DC voltage (TP2) > is 70 mV, but should be around 4.5 V. > > I've got A3 hooked up in isolation on my bench with 5MHz at 1Vrms to J2 > and 600mVrms to J5 at 5.315MHz. Any suggestions for how to troubleshoot > this? I measured Q4 and C12 to ensure they're not shorted - they're > fine. I also checked different points in the circuit and compared them > with the voltage annotations in the schematic. Unfortunately, I'm having > some trouble understanding this circuit, and some of the annotations > confuse me. > > According to the annotations, the base of Q4 (the AGC amplifier), is at > +0.7Vdc for no AGC and +6Vdc for full AGC. But there's no emitter > resistor so how does this get to +6Vdc? Moreover, I don't see a DC path > to supply current to raise the DC voltage. In case it's helpful, Q4 is > apparently a 2N2484. > > Another example is Q9 base is labelled with 2.4V. I think that's dc > since I see Vrms in a number of other places. But I don't see how that > could be true. This is fed via a +20V supply, and the base is biased by > a resistive dividier with 18k and 2k, so it should be at most +2V, but > then there's the parallel path of the BE junction and emitter resistor, > so it would be even lower than this. So, where are they getting 2.4V? > > I checked a number of points in the main signal path and some looked > more or less in line with amplitude annotations, but others didn't. I'm > not totally sure how precise these annotations are supposed to be, > though. > > Finally, I wanted to clarify how the RVFR SRD bias works. The only DC > path to the bias test point is through J4 (which provides the > 60 + 5.315... MHz signal to the SRD). But, the RVFR schematic does not > show any DC bias applied to the SRD. I presume that's there but just not > shown. Is that right? That would make sense, since the pot there (R40) > would divide down the bias. > > Any suggestions would be immensely appreciated. > > Thank you! > Matt > > Matt Huszagh <huszaghmatt@gmail.com> writes: > >> I've decide to start by addressing the cell flooding. Despite the >> incorrect AGC DC voltage, the 60 MHz signal looks roughly ok and from a >> quick look I don't think I can really perform the loop alignment without >> some sort of RVFR signal. >> >> I've attached a current-limited supply at 5V through a 5ohm power >> resistor to the TED wires and have the unit powered with the meter set >> to 2nd harmonic. The meter indication almost immediately started >> deflecting off 0. It's been just a few minutes and the readings are >> getting almost to 10 at max deflection, though deflecting back and >> forth. I'm hopeful this is a good sign. I'll update as this unit has >> more time to de-flood. >> >> Please feel free to make suggestions if I should be following a >> different approach at the moment. >> >> Thanks! >> Matt >> >> Matt Huszagh <huszaghmatt@gmail.com> writes: >> >>> There are a few other things I wanted to mention. >>> >>> I checked the DC voltages around the cell and lamp heaters. J15P3 (cell >>> heater) measured 15.4V and J15P5 measured 14.4V. P4 measured 28.6V. The >>> specified nominal voltages for P3 and P5 are 18V, so these are a bit >>> off, but since the heater currents look ok as measured by the front >>> panel meter, I got the sense that this is probably ok. Thoughts? >>> >>> Also, as part of the pre-emptive maintenance I replaced the lamp >>> assembly film resistor and, in the process, disassembled part of A12 >>> including the lamp assembly. Everything looked fine to my eye and I have >>> pictures I can share if desired. In particular, the lamp assembly was >>> not charred. >>> >>> Matt >>> >>> Matt Huszagh <huszaghmatt@gmail.com> writes: >>> >>>> Hi, >>>> >>>> I have an HP 5065A received in unknown condition. After performing some >>>> pre-emptive maintenance (basically Corby's recommendations here: >>>> https://www.eevblog.com/forum/metrology/inspecting-aligning-and-testing-an-hp-5065a-rubidium/ >>>> and the first aid advice from Luciano here: >>>> http://www.timeok.it/hp5065a-corner-3/) I turned the unit on. +/- 20V >>>> supplies look good and cell and lamp ovens drop down from pegged in >>>> under 45 min to about 32 each. >>>> >>>> However, the 2nd harmonic reading is 0 and of course the unit does not >>>> lock. >>>> >>>> This unit was left unpowered for some time (I got the sense it was at >>>> least several years) before I received it, and therefore cell flooding >>>> is likely. However, I have noticed a few other discrepencies. >>>> >>>> The first is that the A3 TP2 AGC DC voltage measures 71 mV, instead of >>>> the expected >4.5 V. I also checked the 60 MHz signal into 50 ohms and >>>> see about 30 mW. One place in the manual says this should be >1V (which >>>> it is), but another place says 300 mW. Is loading this with 50 ohms >>>> correct? If so, maybe 300 mW is a typo and it should read 30 mW? In any >>>> event the AGC DC voltage is wrong and I suppose needs to be addressed. >>>> >>>> The second issue, and probably a more serious one, is that the PHOTO I >>>> meter reading is 20 (less than the specified 25) and the actual current >>>> measured out of A12 is -10 uA (should be 50 uA or greater). There are >>>> two issues here: (1) the actual current disagrees with the meter >>>> indication and (2) the current is low. Is the low current possibly >>>> explained by cell flooding and or incorrect/insufficient 60 MHz RF >>>> signal application? >>>> >>>> Thoughts on next steps? I believe I need to address both issues >>>> above. Addressing cell flooding is straightforward, but will take some >>>> time. The troubleshooting section for A3 says to perform the loop >>>> adjustment procedure if the AGC DC voltage is wrong. Which should I do >>>> first? Or, should I start with something else? >>>> >>>> Thanks! >>>> Matt