HP 5061: Unable to adjust beam current to less than 46nA

C
cdelect@juno.com
Wed, Jun 7, 2023 5:54 PM

Matt,

You can select a resistor to drop the EM volts down some more.

This would allow you to drop the beam current.

However the instrument will work just fine at your values.

As long as you can adjust the loop gain on the A7 to get 40 on the 2nd
harmonic I'd just leave it as is.

I have a 5061B that has an excellent tube. The minimum E.M. is 1200V and
my beam current is around 48 (mod on). I just use it that way as it's a
hassle to tear into just to change the beam current indication.

Cheers,

Corby

Matt, You can select a resistor to drop the EM volts down some more. This would allow you to drop the beam current. However the instrument will work just fine at your values. As long as you can adjust the loop gain on the A7 to get 40 on the 2nd harmonic I'd just leave it as is. I have a 5061B that has an excellent tube. The minimum E.M. is 1200V and my beam current is around 48 (mod on). I just use it that way as it's a hassle to tear into just to change the beam current indication. Cheers, Corby
AT
Andy Talbot
Wed, Jun 7, 2023 6:18 PM

I found it amusing  in the 5061 manual that they suggest using
the 10Meg input resistance of a Valve Voltmeter as a current shunt to
measure beam current

Andy
www.g4jnt.com

On Wed, 7 Jun 2023 at 19:16, Corby Dawson via time-nuts <
time-nuts@lists.febo.com> wrote:

Matt,

You can select a resistor to drop the EM volts down some more.

This would allow you to drop the beam current.

However the instrument will work just fine at your values.

As long as you can adjust the loop gain on the A7 to get 40 on the 2nd
harmonic I'd just leave it as is.

I have a 5061B that has an excellent tube. The minimum E.M. is 1200V and
my beam current is around 48 (mod on). I just use it that way as it's a
hassle to tear into just to change the beam current indication.

Cheers,

Corby


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

I found it amusing in the 5061 manual that they suggest using the 10Meg input resistance of a Valve Voltmeter as a current shunt to measure beam current Andy www.g4jnt.com On Wed, 7 Jun 2023 at 19:16, Corby Dawson via time-nuts < time-nuts@lists.febo.com> wrote: > Matt, > > You can select a resistor to drop the EM volts down some more. > > This would allow you to drop the beam current. > > However the instrument will work just fine at your values. > > As long as you can adjust the loop gain on the A7 to get 40 on the 2nd > harmonic I'd just leave it as is. > > I have a 5061B that has an excellent tube. The minimum E.M. is 1200V and > my beam current is around 48 (mod on). I just use it that way as it's a > hassle to tear into just to change the beam current indication. > > Cheers, > > Corby > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com >
MH
Matt Huszagh
Wed, Jun 7, 2023 7:09 PM

Corby Dawson via time-nuts time-nuts@lists.febo.com writes:

You can select a resistor to drop the EM volts down some more.

This would allow you to drop the beam current.

However the instrument will work just fine at your values.

As long as you can adjust the loop gain on the A7 to get 40 on the 2nd
harmonic I'd just leave it as is.

I have a 5061B that has an excellent tube. The minimum E.M. is 1200V and
my beam current is around 48 (mod on). I just use it that way as it's a
hassle to tear into just to change the beam current indication.

Thanks for the confirmation Corby.

I can adjust the loop gain to get a 2nd harmonic reading of 40 at the
lowest beam current setting. However, I performed the loop gain
adjustment (5-220 in manual) to set the A8J1 phase detector test output
to +/- 1.6V when the fine frequency adjust is set to 200 and 300
(normally 250, where the test output is 0). After doing this, the 2nd
harmonic reading is >50. When I adjust the loop gain such that the 2nd
harmonic is at 40, the phase det test output is much less than +/-
1.6V. I suppose maybe that +/-1.6V is only applicable with a 25nA beam
current. Is that right?  Should I neglect the phase det test output
setting of 1.6V and just set the loop gain such that the 2nd harmonic
reads 40?

Matt

Corby Dawson via time-nuts <time-nuts@lists.febo.com> writes: > You can select a resistor to drop the EM volts down some more. > > This would allow you to drop the beam current. > > However the instrument will work just fine at your values. > > As long as you can adjust the loop gain on the A7 to get 40 on the 2nd > harmonic I'd just leave it as is. > > I have a 5061B that has an excellent tube. The minimum E.M. is 1200V and > my beam current is around 48 (mod on). I just use it that way as it's a > hassle to tear into just to change the beam current indication. Thanks for the confirmation Corby. I can adjust the loop gain to get a 2nd harmonic reading of 40 at the lowest beam current setting. However, I performed the loop gain adjustment (5-220 in manual) to set the A8J1 phase detector test output to +/- 1.6V when the fine frequency adjust is set to 200 and 300 (normally 250, where the test output is 0). After doing this, the 2nd harmonic reading is >50. When I adjust the loop gain such that the 2nd harmonic is at 40, the phase det test output is much less than +/- 1.6V. I suppose maybe that +/-1.6V is only applicable with a 25nA beam current. Is that right? Should I neglect the phase det test output setting of 1.6V and just set the loop gain such that the 2nd harmonic reads 40? Matt