time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

Z3805 initial behaviour after power up

UB
Ulrich Bangert
Wed, Jun 17, 2009 8:06 AM

Gents,

yesterday I received the Z3805 that I had bought from fluke.I on eBay and
immediately put it to work. I recorded the EFC value as well as the PPS TI
value over about 20 hours. Attached is the graph of thes two values.

I interprete this graph in this way:

  1. Immediately after power up (perhaps up to the end of the site survey) the
    regulation loop seems to work with a very short time constant which is able
    to bring the PPS TI close to zero although the OCXO has lots of drift at
    this time.

  2. Then the loop time constant seems to switch to a significant higher value
    which does not keep up with the oscillator's drift, which leads to the
    negative excursion of nearly 500 ns.

  3. This seems to be too much for the regulation which leads to switching
    back to the short time constant and to bring the PPS TI back to near zero
    and to return to the longer time constant again...

  4. ...which lets the procedure repeat again and again: Drift to -500 ns and
    pullback.

  5. Note that I thought it was a kind of malfunction and send a
    :system:preset command to the device after the last negative PPS TI spike.
    The result was, that the total procedure repeated: About 2 hours of PPS TI
    close to zero then again the drift and pullback procedure however with a
    reduced drift rate.

  6. After 11 hours or so the regulation loop's time constant seems to match
    the OCXO's drift and a "real" lock of the pll seems to take place. (The lock
    led had been on already hours ago...)

Since I do not have an manual for the Z3805: Can any of you Z3801 owners
second this behaviour to be normal or would you think that it is kind of
special for the Z3805 or even a real malfunction (in the sense that the
OCXO's drift is above the specs in the beginning)?

Best regards
Ulrich Bangert

Ulrich Bangert
www.ulrich-bangert.de
Ortholzer Weg 1
27243 Gross Ippener

Gents, yesterday I received the Z3805 that I had bought from fluke.I on eBay and immediately put it to work. I recorded the EFC value as well as the PPS TI value over about 20 hours. Attached is the graph of thes two values. I interprete this graph in this way: 1) Immediately after power up (perhaps up to the end of the site survey) the regulation loop seems to work with a very short time constant which is able to bring the PPS TI close to zero although the OCXO has lots of drift at this time. 2) Then the loop time constant seems to switch to a significant higher value which does not keep up with the oscillator's drift, which leads to the negative excursion of nearly 500 ns. 3) This seems to be too much for the regulation which leads to switching back to the short time constant and to bring the PPS TI back to near zero and to return to the longer time constant again... 4) ...which lets the procedure repeat again and again: Drift to -500 ns and pullback. 5) Note that I thought it was a kind of malfunction and send a :system:preset command to the device after the last negative PPS TI spike. The result was, that the total procedure repeated: About 2 hours of PPS TI close to zero then again the drift and pullback procedure however with a reduced drift rate. 6) After 11 hours or so the regulation loop's time constant seems to match the OCXO's drift and a "real" lock of the pll seems to take place. (The lock led had been on already hours ago...) Since I do not have an manual for the Z3805: Can any of you Z3801 owners second this behaviour to be normal or would you think that it is kind of special for the Z3805 or even a real malfunction (in the sense that the OCXO's drift is above the specs in the beginning)? Best regards Ulrich Bangert Ulrich Bangert www.ulrich-bangert.de Ortholzer Weg 1 27243 Gross Ippener
JM
John Miles
Wed, Jun 17, 2009 8:25 AM

Since I do not have an manual for the Z3805: Can any of you Z3801 owners
second this behaviour to be normal or would you think that it is kind of
special for the Z3805 or even a real malfunction (in the sense that the
OCXO's drift is above the specs in the beginning)?

Any OCXO that has been sitting around for years is likely to undergo rapid
aging when you first bring it up.  It will be so far out of spec that I
wouldn't even bother looking at the loop behavior until it's been running
for a week or two.  Chances are it'll be fine by then.

-- john, KE5FX

> Since I do not have an manual for the Z3805: Can any of you Z3801 owners > second this behaviour to be normal or would you think that it is kind of > special for the Z3805 or even a real malfunction (in the sense that the > OCXO's drift is above the specs in the beginning)? Any OCXO that has been sitting around for years is likely to undergo rapid aging when you first bring it up. It will be so far out of spec that I wouldn't even bother looking at the loop behavior until it's been running for a week or two. Chances are it'll be fine by then. -- john, KE5FX
UB
Ulrich Bangert
Wed, Jun 17, 2009 8:52 AM

John,

Any OCXO that has been sitting around for years is likely to
undergo rapid aging when you first bring it up.

Sure! I had better formulated my question like: A constructor of a GPSDO
knows, that any OCXO that has been sitting around for years is likely to
undergo rapid aging when it is first brouht up. When Symmetricon choosed to
switch to a long time constant early after power up, were they perhaps
expecting a lower OCXO drift than I am observing now?

73s Ulrich DF6JB

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von John Miles
Gesendet: Mittwoch, 17. Juni 2009 10:25
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Since I do not have an manual for the Z3805: Can any of you Z3801
owners second this behaviour to be normal or would you

think that it

is kind of special for the Z3805 or even a real malfunction (in the
sense that the OCXO's drift is above the specs in the beginning)?

Any OCXO that has been sitting around for years is likely to
undergo rapid aging when you first bring it up.  It will be
so far out of spec that I wouldn't even bother looking at the
loop behavior until it's been running for a week or two.
Chances are it'll be fine by then.

-- john, KE5FX


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

John, > Any OCXO that has been sitting around for years is likely to > undergo rapid aging when you first bring it up. Sure! I had better formulated my question like: A constructor of a GPSDO knows, that any OCXO that has been sitting around for years is likely to undergo rapid aging when it is first brouht up. When Symmetricon choosed to switch to a long time constant early after power up, were they perhaps expecting a lower OCXO drift than I am observing now? 73s Ulrich DF6JB > -----Ursprungliche Nachricht----- > Von: time-nuts-bounces@febo.com > [mailto:time-nuts-bounces@febo.com] Im Auftrag von John Miles > Gesendet: Mittwoch, 17. Juni 2009 10:25 > An: Discussion of precise time and frequency measurement > Betreff: Re: [time-nuts] Z3805 initial behaviour after power up > > > > > Since I do not have an manual for the Z3805: Can any of you Z3801 > > owners second this behaviour to be normal or would you > think that it > > is kind of special for the Z3805 or even a real malfunction (in the > > sense that the OCXO's drift is above the specs in the beginning)? > > Any OCXO that has been sitting around for years is likely to > undergo rapid aging when you first bring it up. It will be > so far out of spec that I wouldn't even bother looking at the > loop behavior until it's been running for a week or two. > Chances are it'll be fine by then. > > -- john, KE5FX > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
SR
Steve Rooke
Wed, Jun 17, 2009 9:03 AM

Ulrich,

What are you using to produce the graph you attached please?

73,
Steve

2009/6/17 Ulrich Bangert df6jb@ulrich-bangert.de:

Gents,

yesterday I received the Z3805 that I had bought from fluke.I on eBay and
immediately put it to work. I recorded the EFC value as well as the PPS TI
value over about 20 hours. Attached is the graph of thes two values.

I interprete this graph in this way:

  1. Immediately after power up (perhaps up to the end of the site survey) the
    regulation loop seems to work with a very short time constant which is able
    to bring the PPS TI close to zero although the OCXO has lots of drift at
    this time.

  2. Then the loop time constant seems to switch to a significant higher value
    which does not keep up with the oscillator's drift, which leads to the
    negative excursion of nearly 500 ns.

  3. This seems to be too much for the regulation which leads to switching
    back to the short time constant and to bring the PPS TI back to near zero
    and to return to the longer time constant again...

  4. ...which lets the procedure repeat again and again: Drift to -500 ns and
    pullback.

  5. Note that I thought it was a kind of malfunction and send a
    :system:preset command to the device after the last negative PPS TI spike.
    The result was, that the total procedure repeated: About 2 hours of PPS TI
    close to zero then again the drift and pullback procedure however with a
    reduced drift rate.

  6. After 11 hours or so the regulation loop's time constant seems to match
    the OCXO's drift and a "real" lock of the pll seems to take place. (The lock
    led had been on already hours ago...)

Since I do not have an manual for the Z3805: Can any of you Z3801 owners
second this behaviour to be normal or would you think that it is kind of
special for the Z3805 or even a real malfunction (in the sense that the
OCXO's drift is above the specs in the beginning)?

Best regards
Ulrich Bangert

Ulrich Bangert
www.ulrich-bangert.de
Ortholzer Weg 1
27243 Gross Ippener


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
Steve Rooke - ZL3TUV & G8KVD & JAKDTTNW
A man with one clock knows what time it is;
A man with two clocks is never quite sure.

Ulrich, What are you using to produce the graph you attached please? 73, Steve 2009/6/17 Ulrich Bangert <df6jb@ulrich-bangert.de>: > Gents, > > yesterday I received the Z3805 that I had bought from fluke.I on eBay and > immediately put it to work. I recorded the EFC value as well as the PPS TI > value over about 20 hours. Attached is the graph of thes two values. > > I interprete this graph in this way: > > 1) Immediately after power up (perhaps up to the end of the site survey) the > regulation loop seems to work with a very short time constant which is able > to bring the PPS TI close to zero although the OCXO has lots of drift at > this time. > > 2) Then the loop time constant seems to switch to a significant higher value > which does not keep up with the oscillator's drift, which leads to the > negative excursion of nearly 500 ns. > > 3) This seems to be too much for the regulation which leads to switching > back to the short time constant and to bring the PPS TI back to near zero > and to return to the longer time constant again... > > 4) ...which lets the procedure repeat again and again: Drift to -500 ns and > pullback. > > 5) Note that I thought it was a kind of malfunction and send a > :system:preset command to the device after the last negative PPS TI spike. > The result was, that the total procedure repeated: About 2 hours of PPS TI > close to zero then again the drift and pullback procedure however with a > reduced drift rate. > > 6) After 11 hours or so the regulation loop's time constant seems to match > the OCXO's drift and a "real" lock of the pll seems to take place. (The lock > led had been on already hours ago...) > > Since I do not have an manual for the Z3805: Can any of you Z3801 owners > second this behaviour to be normal or would you think that it is kind of > special for the Z3805 or even a real malfunction (in the sense that the > OCXO's drift is above the specs in the beginning)? > > Best regards > Ulrich Bangert > > Ulrich Bangert > www.ulrich-bangert.de > Ortholzer Weg 1 > 27243 Gross Ippener > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > -- Steve Rooke - ZL3TUV & G8KVD & JAKDTTNW A man with one clock knows what time it is; A man with two clocks is never quite sure.
UB
Ulrich Bangert
Wed, Jun 17, 2009 9:07 AM

Steve,

my PLOTTER utility. I thought you did already try it out?

73s Ulrich, DF6JB

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Steve Rooke
Gesendet: Mittwoch, 17. Juni 2009 11:03
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Ulrich,

What are you using to produce the graph you attached please?

73,
Steve

2009/6/17 Ulrich Bangert df6jb@ulrich-bangert.de:

Gents,

yesterday I received the Z3805 that I had bought from

fluke.I on eBay

and immediately put it to work. I recorded the EFC value as well as
the PPS TI value over about 20 hours. Attached is the graph of thes
two values.

I interprete this graph in this way:

  1. Immediately after power up (perhaps up to the end of the site
    survey) the regulation loop seems to work with a very short time
    constant which is able to bring the PPS TI close to zero

although the

OCXO has lots of drift at this time.

  1. Then the loop time constant seems to switch to a

significant higher

value which does not keep up with the oscillator's drift,

which leads

to the negative excursion of nearly 500 ns.

  1. This seems to be too much for the regulation which leads to
    switching back to the short time constant and to bring the

PPS TI back

to near zero and to return to the longer time constant again...

  1. ...which lets the procedure repeat again and again:

Drift to -500

ns and pullback.

  1. Note that I thought it was a kind of malfunction and send a
    :system:preset command to the device after the last negative PPS TI
    spike. The result was, that the total procedure repeated: About 2
    hours of PPS TI close to zero then again the drift and pullback
    procedure however with a reduced drift rate.

  2. After 11 hours or so the regulation loop's time constant

seems to

match the OCXO's drift and a "real" lock of the pll seems to take
place. (The lock led had been on already hours ago...)

Since I do not have an manual for the Z3805: Can any of you Z3801
owners second this behaviour to be normal or would you

think that it

is kind of special for the Z3805 or even a real malfunction (in the
sense that the OCXO's drift is above the specs in the beginning)?

Best regards
Ulrich Bangert

Ulrich Bangert
www.ulrich-bangert.de
Ortholzer Weg 1
27243 Gross Ippener


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
Steve Rooke - ZL3TUV & G8KVD & JAKDTTNW
A man with one clock knows what time it is;
A man with two clocks is never quite sure.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Steve, my PLOTTER utility. I thought you did already try it out? 73s Ulrich, DF6JB > -----Ursprungliche Nachricht----- > Von: time-nuts-bounces@febo.com > [mailto:time-nuts-bounces@febo.com] Im Auftrag von Steve Rooke > Gesendet: Mittwoch, 17. Juni 2009 11:03 > An: Discussion of precise time and frequency measurement > Betreff: Re: [time-nuts] Z3805 initial behaviour after power up > > > Ulrich, > > What are you using to produce the graph you attached please? > > 73, > Steve > > 2009/6/17 Ulrich Bangert <df6jb@ulrich-bangert.de>: > > Gents, > > > > yesterday I received the Z3805 that I had bought from > fluke.I on eBay > > and immediately put it to work. I recorded the EFC value as well as > > the PPS TI value over about 20 hours. Attached is the graph of thes > > two values. > > > > I interprete this graph in this way: > > > > 1) Immediately after power up (perhaps up to the end of the site > > survey) the regulation loop seems to work with a very short time > > constant which is able to bring the PPS TI close to zero > although the > > OCXO has lots of drift at this time. > > > > 2) Then the loop time constant seems to switch to a > significant higher > > value which does not keep up with the oscillator's drift, > which leads > > to the negative excursion of nearly 500 ns. > > > > 3) This seems to be too much for the regulation which leads to > > switching back to the short time constant and to bring the > PPS TI back > > to near zero and to return to the longer time constant again... > > > > 4) ...which lets the procedure repeat again and again: > Drift to -500 > > ns and pullback. > > > > 5) Note that I thought it was a kind of malfunction and send a > > :system:preset command to the device after the last negative PPS TI > > spike. The result was, that the total procedure repeated: About 2 > > hours of PPS TI close to zero then again the drift and pullback > > procedure however with a reduced drift rate. > > > > 6) After 11 hours or so the regulation loop's time constant > seems to > > match the OCXO's drift and a "real" lock of the pll seems to take > > place. (The lock led had been on already hours ago...) > > > > Since I do not have an manual for the Z3805: Can any of you Z3801 > > owners second this behaviour to be normal or would you > think that it > > is kind of special for the Z3805 or even a real malfunction (in the > > sense that the OCXO's drift is above the specs in the beginning)? > > > > Best regards > > Ulrich Bangert > > > > Ulrich Bangert > > www.ulrich-bangert.de > > Ortholzer Weg 1 > > 27243 Gross Ippener > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > > > > > -- > Steve Rooke - ZL3TUV & G8KVD & JAKDTTNW > A man with one clock knows what time it is; > A man with two clocks is never quite sure. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
SR
Steve Rooke
Wed, Jun 17, 2009 9:28 AM

Ulrich,

I must have lost the plot here but I thought that was only for HPIB
data. How are you collecting the data?

73,
Steve

2009/6/17 Ulrich Bangert df6jb@ulrich-bangert.de:

Steve,

my PLOTTER utility. I thought you did already try it out?

73s Ulrich, DF6JB

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Steve Rooke
Gesendet: Mittwoch, 17. Juni 2009 11:03
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Ulrich,

What are you using to produce the graph you attached please?

73,
Steve

2009/6/17 Ulrich Bangert df6jb@ulrich-bangert.de:

Gents,

yesterday I received the Z3805 that I had bought from

fluke.I on eBay

and immediately put it to work. I recorded the EFC value as well as
the PPS TI value over about 20 hours. Attached is the graph of thes
two values.

I interprete this graph in this way:

  1. Immediately after power up (perhaps up to the end of the site
    survey) the regulation loop seems to work with a very short time
    constant which is able to bring the PPS TI close to zero

although the

OCXO has lots of drift at this time.

  1. Then the loop time constant seems to switch to a

significant higher

value which does not keep up with the oscillator's drift,

which leads

to the negative excursion of nearly 500 ns.

  1. This seems to be too much for the regulation which leads to
    switching back to the short time constant and to bring the

PPS TI back

to near zero and to return to the longer time constant again...

  1. ...which lets the procedure repeat again and again:

Drift to -500

ns and pullback.

  1. Note that I thought it was a kind of malfunction and send a
    :system:preset command to the device after the last negative PPS TI
    spike. The result was, that the total procedure repeated: About 2
    hours of PPS TI close to zero then again the drift and pullback
    procedure however with a reduced drift rate.

  2. After 11 hours or so the regulation loop's time constant

seems to

match the OCXO's drift and a "real" lock of the pll seems to take
place. (The lock led had been on already hours ago...)

Since I do not have an manual for the Z3805: Can any of you Z3801
owners second this behaviour to be normal or would you

think that it

is kind of special for the Z3805 or even a real malfunction (in the
sense that the OCXO's drift is above the specs in the beginning)?

Best regards
Ulrich Bangert

Ulrich Bangert
www.ulrich-bangert.de
Ortholzer Weg 1
27243 Gross Ippener


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
Steve Rooke - ZL3TUV & G8KVD & JAKDTTNW
A man with one clock knows what time it is;
A man with two clocks is never quite sure.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
Steve Rooke - ZL3TUV & G8KVD & JAKDTTNW
A man with one clock knows what time it is;
A man with two clocks is never quite sure.

Ulrich, I must have lost the plot here but I thought that was only for HPIB data. How are you collecting the data? 73, Steve 2009/6/17 Ulrich Bangert <df6jb@ulrich-bangert.de>: > Steve, > > my PLOTTER utility. I thought you did already try it out? > > 73s Ulrich, DF6JB > >> -----Ursprungliche Nachricht----- >> Von: time-nuts-bounces@febo.com >> [mailto:time-nuts-bounces@febo.com] Im Auftrag von Steve Rooke >> Gesendet: Mittwoch, 17. Juni 2009 11:03 >> An: Discussion of precise time and frequency measurement >> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up >> >> >> Ulrich, >> >> What are you using to produce the graph you attached please? >> >> 73, >> Steve >> >> 2009/6/17 Ulrich Bangert <df6jb@ulrich-bangert.de>: >> > Gents, >> > >> > yesterday I received the Z3805 that I had bought from >> fluke.I on eBay >> > and immediately put it to work. I recorded the EFC value as well as >> > the PPS TI value over about 20 hours. Attached is the graph of thes >> > two values. >> > >> > I interprete this graph in this way: >> > >> > 1) Immediately after power up (perhaps up to the end of the site >> > survey) the regulation loop seems to work with a very short time >> > constant which is able to bring the PPS TI close to zero >> although the >> > OCXO has lots of drift at this time. >> > >> > 2) Then the loop time constant seems to switch to a >> significant higher >> > value which does not keep up with the oscillator's drift, >> which leads >> > to the negative excursion of nearly 500 ns. >> > >> > 3) This seems to be too much for the regulation which leads to >> > switching back to the short time constant and to bring the >> PPS TI back >> > to near zero and to return to the longer time constant again... >> > >> > 4) ...which lets the procedure repeat again and again: >> Drift to -500 >> > ns and pullback. >> > >> > 5) Note that I thought it was a kind of malfunction and send a >> > :system:preset command to the device after the last negative PPS TI >> > spike. The result was, that the total procedure repeated: About 2 >> > hours of PPS TI close to zero then again the drift and pullback >> > procedure however with a reduced drift rate. >> > >> > 6) After 11 hours or so the regulation loop's time constant >> seems to >> > match the OCXO's drift and a "real" lock of the pll seems to take >> > place. (The lock led had been on already hours ago...) >> > >> > Since I do not have an manual for the Z3805: Can any of you Z3801 >> > owners second this behaviour to be normal or would you >> think that it >> > is kind of special for the Z3805 or even a real malfunction (in the >> > sense that the OCXO's drift is above the specs in the beginning)? >> > >> > Best regards >> > Ulrich Bangert >> > >> > Ulrich Bangert >> > www.ulrich-bangert.de >> > Ortholzer Weg 1 >> > 27243 Gross Ippener >> > >> > _______________________________________________ >> > time-nuts mailing list -- time-nuts@febo.com >> > To unsubscribe, go to >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> > and follow the instructions there. >> > >> >> >> >> -- >> Steve Rooke - ZL3TUV & G8KVD & JAKDTTNW >> A man with one clock knows what time it is; >> A man with two clocks is never quite sure. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > -- Steve Rooke - ZL3TUV & G8KVD & JAKDTTNW A man with one clock knows what time it is; A man with two clocks is never quite sure.
BG
Bruce Griffiths
Wed, Jun 17, 2009 10:39 AM

Steve Rooke wrote:

Ulrich,

I must have lost the plot here but I thought that was only for HPIB
data. How are you collecting the data?

73,
Steve

2009/6/17 Ulrich Bangert df6jb@ulrich-bangert.de:

Steve,

my PLOTTER utility. I thought you did already try it out?

73s Ulrich, DF6JB

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Steve Rooke
Gesendet: Mittwoch, 17. Juni 2009 11:03
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Ulrich,

What are you using to produce the graph you attached please?

73,
Steve

2009/6/17 Ulrich Bangert df6jb@ulrich-bangert.de:

Gents,

yesterday I received the Z3805 that I had bought from

fluke.I on eBay

and immediately put it to work. I recorded the EFC value as well as
the PPS TI value over about 20 hours. Attached is the graph of thes
two values.

I interprete this graph in this way:

  1. Immediately after power up (perhaps up to the end of the site
    survey) the regulation loop seems to work with a very short time
    constant which is able to bring the PPS TI close to zero

although the

OCXO has lots of drift at this time.

  1. Then the loop time constant seems to switch to a

significant higher

value which does not keep up with the oscillator's drift,

which leads

to the negative excursion of nearly 500 ns.

  1. This seems to be too much for the regulation which leads to
    switching back to the short time constant and to bring the

PPS TI back

to near zero and to return to the longer time constant again...

  1. ...which lets the procedure repeat again and again:

Drift to -500

ns and pullback.

  1. Note that I thought it was a kind of malfunction and send a
    :system:preset command to the device after the last negative PPS TI
    spike. The result was, that the total procedure repeated: About 2
    hours of PPS TI close to zero then again the drift and pullback
    procedure however with a reduced drift rate.

  2. After 11 hours or so the regulation loop's time constant

seems to

match the OCXO's drift and a "real" lock of the pll seems to take
place. (The lock led had been on already hours ago...)

Since I do not have an manual for the Z3805: Can any of you Z3801
owners second this behaviour to be normal or would you

think that it

is kind of special for the Z3805 or even a real malfunction (in the
sense that the OCXO's drift is above the specs in the beginning)?

Best regards
Ulrich Bangert

Ulrich Bangert
www.ulrich-bangert.de
Ortholzer Weg 1
27243 Gross Ippener


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
Steve Rooke - ZL3TUV & G8KVD & JAKDTTNW
A man with one clock knows what time it is;
A man with two clocks is never quite sure.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Steve

You're confusing Ulrich's PLOTTER program with his EZGPIB program (which
can also communicate with RS232 devices like GPSDOs).
Once the data has been logged using EZGPIB, then Plotter can be used to
plot/analyse the stored results.

Bruce

Steve Rooke wrote: > Ulrich, > > I must have lost the plot here but I thought that was only for HPIB > data. How are you collecting the data? > > 73, > Steve > > 2009/6/17 Ulrich Bangert <df6jb@ulrich-bangert.de>: > >> Steve, >> >> my PLOTTER utility. I thought you did already try it out? >> >> 73s Ulrich, DF6JB >> >> >>> -----Ursprungliche Nachricht----- >>> Von: time-nuts-bounces@febo.com >>> [mailto:time-nuts-bounces@febo.com] Im Auftrag von Steve Rooke >>> Gesendet: Mittwoch, 17. Juni 2009 11:03 >>> An: Discussion of precise time and frequency measurement >>> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up >>> >>> >>> Ulrich, >>> >>> What are you using to produce the graph you attached please? >>> >>> 73, >>> Steve >>> >>> 2009/6/17 Ulrich Bangert <df6jb@ulrich-bangert.de>: >>> >>>> Gents, >>>> >>>> yesterday I received the Z3805 that I had bought from >>>> >>> fluke.I on eBay >>> >>>> and immediately put it to work. I recorded the EFC value as well as >>>> the PPS TI value over about 20 hours. Attached is the graph of thes >>>> two values. >>>> >>>> I interprete this graph in this way: >>>> >>>> 1) Immediately after power up (perhaps up to the end of the site >>>> survey) the regulation loop seems to work with a very short time >>>> constant which is able to bring the PPS TI close to zero >>>> >>> although the >>> >>>> OCXO has lots of drift at this time. >>>> >>>> 2) Then the loop time constant seems to switch to a >>>> >>> significant higher >>> >>>> value which does not keep up with the oscillator's drift, >>>> >>> which leads >>> >>>> to the negative excursion of nearly 500 ns. >>>> >>>> 3) This seems to be too much for the regulation which leads to >>>> switching back to the short time constant and to bring the >>>> >>> PPS TI back >>> >>>> to near zero and to return to the longer time constant again... >>>> >>>> 4) ...which lets the procedure repeat again and again: >>>> >>> Drift to -500 >>> >>>> ns and pullback. >>>> >>>> 5) Note that I thought it was a kind of malfunction and send a >>>> :system:preset command to the device after the last negative PPS TI >>>> spike. The result was, that the total procedure repeated: About 2 >>>> hours of PPS TI close to zero then again the drift and pullback >>>> procedure however with a reduced drift rate. >>>> >>>> 6) After 11 hours or so the regulation loop's time constant >>>> >>> seems to >>> >>>> match the OCXO's drift and a "real" lock of the pll seems to take >>>> place. (The lock led had been on already hours ago...) >>>> >>>> Since I do not have an manual for the Z3805: Can any of you Z3801 >>>> owners second this behaviour to be normal or would you >>>> >>> think that it >>> >>>> is kind of special for the Z3805 or even a real malfunction (in the >>>> sense that the OCXO's drift is above the specs in the beginning)? >>>> >>>> Best regards >>>> Ulrich Bangert >>>> >>>> Ulrich Bangert >>>> www.ulrich-bangert.de >>>> Ortholzer Weg 1 >>>> 27243 Gross Ippener >>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com >>>> To unsubscribe, go to >>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>> and follow the instructions there. >>>> >>>> >>> >>> -- >>> Steve Rooke - ZL3TUV & G8KVD & JAKDTTNW >>> A man with one clock knows what time it is; >>> A man with two clocks is never quite sure. >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> >> > > > > Steve You're confusing Ulrich's PLOTTER program with his EZGPIB program (which can also communicate with RS232 devices like GPSDOs). Once the data has been logged using EZGPIB, then Plotter can be used to plot/analyse the stored results. Bruce
UB
Ulrich Bangert
Wed, Jun 17, 2009 11:33 AM

Steve,

I must have lost the plot here but I thought that was only
for HPIB data.

By far not! PLOTTER does a lot to everything than can be pressed into ASCII
data column files.

How are you collecting the data?

That has been done with a quick and dirty script for my EZGPIB utility
(which handles serial communication and some other goodies as well). I have
just included this script as "Z3805.488" to the examples.

Don't accuse me that I did not have warned you: PLOTTER is written in the
P-language while EZGPIB (which is written in the P-language itself) executes
scripts in the P-language, P-square if you like. Better get on rubber gloves
before you touch anything of it, hi.

73s Ulrich, DF6JB

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Steve Rooke
Gesendet: Mittwoch, 17. Juni 2009 11:29
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Ulrich,

I must have lost the plot here but I thought that was only
for HPIB data. How are you collecting the data?

73,
Steve

2009/6/17 Ulrich Bangert df6jb@ulrich-bangert.de:

Steve,

my PLOTTER utility. I thought you did already try it out?

73s Ulrich, DF6JB

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com

Im Auftrag von Steve Rooke
Gesendet: Mittwoch, 17. Juni 2009 11:03
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Ulrich,

What are you using to produce the graph you attached please?

73,
Steve

2009/6/17 Ulrich Bangert df6jb@ulrich-bangert.de:

Gents,

yesterday I received the Z3805 that I had bought from

fluke.I on eBay

and immediately put it to work. I recorded the EFC value

as well as

the PPS TI value over about 20 hours. Attached is the

graph of thes

two values.

I interprete this graph in this way:

  1. Immediately after power up (perhaps up to the end of the site
    survey) the regulation loop seems to work with a very short time
    constant which is able to bring the PPS TI close to zero

although the

OCXO has lots of drift at this time.

  1. Then the loop time constant seems to switch to a

significant higher

value which does not keep up with the oscillator's drift,

which leads

to the negative excursion of nearly 500 ns.

  1. This seems to be too much for the regulation which leads to
    switching back to the short time constant and to bring the

PPS TI back

to near zero and to return to the longer time constant again...

  1. ...which lets the procedure repeat again and again:

Drift to -500

ns and pullback.

  1. Note that I thought it was a kind of malfunction and send a
    :system:preset command to the device after the last

negative PPS TI

spike. The result was, that the total procedure

repeated: About 2

hours of PPS TI close to zero then again the drift and pullback
procedure however with a reduced drift rate.

  1. After 11 hours or so the regulation loop's time constant

seems to

match the OCXO's drift and a "real" lock of the pll

seems to take

place. (The lock led had been on already hours ago...)

Since I do not have an manual for the Z3805: Can any of

you Z3801

owners second this behaviour to be normal or would you

think that it

is kind of special for the Z3805 or even a real

malfunction (in the

sense that the OCXO's drift is above the specs in the beginning)?

Best regards
Ulrich Bangert

Ulrich Bangert
www.ulrich-bangert.de
Ortholzer Weg 1
27243 Gross Ippener


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
Steve Rooke - ZL3TUV & G8KVD & JAKDTTNW
A man with one clock knows what time it is;
A man with two clocks is never quite sure.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
Steve Rooke - ZL3TUV & G8KVD & JAKDTTNW
A man with one clock knows what time it is;
A man with two clocks is never quite sure.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Steve, > I must have lost the plot here but I thought that was only > for HPIB data. By far not! PLOTTER does a lot to everything than can be pressed into ASCII data column files. > How are you collecting the data? That has been done with a quick and dirty script for my EZGPIB utility (which handles serial communication and some other goodies as well). I have just included this script as "Z3805.488" to the examples. Don't accuse me that I did not have warned you: PLOTTER is written in the P-language while EZGPIB (which is written in the P-language itself) executes scripts in the P-language, P-square if you like. Better get on rubber gloves before you touch anything of it, hi. 73s Ulrich, DF6JB > -----Ursprungliche Nachricht----- > Von: time-nuts-bounces@febo.com > [mailto:time-nuts-bounces@febo.com] Im Auftrag von Steve Rooke > Gesendet: Mittwoch, 17. Juni 2009 11:29 > An: Discussion of precise time and frequency measurement > Betreff: Re: [time-nuts] Z3805 initial behaviour after power up > > > Ulrich, > > I must have lost the plot here but I thought that was only > for HPIB data. How are you collecting the data? > > 73, > Steve > > 2009/6/17 Ulrich Bangert <df6jb@ulrich-bangert.de>: > > Steve, > > > > my PLOTTER utility. I thought you did already try it out? > > > > 73s Ulrich, DF6JB > > > >> -----Ursprungliche Nachricht----- > >> Von: time-nuts-bounces@febo.com > [mailto:time-nuts-bounces@febo.com] > >> Im Auftrag von Steve Rooke > >> Gesendet: Mittwoch, 17. Juni 2009 11:03 > >> An: Discussion of precise time and frequency measurement > >> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up > >> > >> > >> Ulrich, > >> > >> What are you using to produce the graph you attached please? > >> > >> 73, > >> Steve > >> > >> 2009/6/17 Ulrich Bangert <df6jb@ulrich-bangert.de>: > >> > Gents, > >> > > >> > yesterday I received the Z3805 that I had bought from > >> fluke.I on eBay > >> > and immediately put it to work. I recorded the EFC value > as well as > >> > the PPS TI value over about 20 hours. Attached is the > graph of thes > >> > two values. > >> > > >> > I interprete this graph in this way: > >> > > >> > 1) Immediately after power up (perhaps up to the end of the site > >> > survey) the regulation loop seems to work with a very short time > >> > constant which is able to bring the PPS TI close to zero > >> although the > >> > OCXO has lots of drift at this time. > >> > > >> > 2) Then the loop time constant seems to switch to a > >> significant higher > >> > value which does not keep up with the oscillator's drift, > >> which leads > >> > to the negative excursion of nearly 500 ns. > >> > > >> > 3) This seems to be too much for the regulation which leads to > >> > switching back to the short time constant and to bring the > >> PPS TI back > >> > to near zero and to return to the longer time constant again... > >> > > >> > 4) ...which lets the procedure repeat again and again: > >> Drift to -500 > >> > ns and pullback. > >> > > >> > 5) Note that I thought it was a kind of malfunction and send a > >> > :system:preset command to the device after the last > negative PPS TI > >> > spike. The result was, that the total procedure > repeated: About 2 > >> > hours of PPS TI close to zero then again the drift and pullback > >> > procedure however with a reduced drift rate. > >> > > >> > 6) After 11 hours or so the regulation loop's time constant > >> seems to > >> > match the OCXO's drift and a "real" lock of the pll > seems to take > >> > place. (The lock led had been on already hours ago...) > >> > > >> > Since I do not have an manual for the Z3805: Can any of > you Z3801 > >> > owners second this behaviour to be normal or would you > >> think that it > >> > is kind of special for the Z3805 or even a real > malfunction (in the > >> > sense that the OCXO's drift is above the specs in the beginning)? > >> > > >> > Best regards > >> > Ulrich Bangert > >> > > >> > Ulrich Bangert > >> > www.ulrich-bangert.de > >> > Ortholzer Weg 1 > >> > 27243 Gross Ippener > >> > > >> > _______________________________________________ > >> > time-nuts mailing list -- time-nuts@febo.com > >> > To unsubscribe, go to > >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> > and follow the instructions there. > >> > > >> > >> > >> > >> -- > >> Steve Rooke - ZL3TUV & G8KVD & JAKDTTNW > >> A man with one clock knows what time it is; > >> A man with two clocks is never quite sure. > >> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@febo.com > >> To unsubscribe, go to > >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> and follow the instructions there. > > > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > > > > > -- > Steve Rooke - ZL3TUV & G8KVD & JAKDTTNW > A man with one clock knows what time it is; > A man with two clocks is never quite sure. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BK
Brian Kirby
Wed, Jun 17, 2009 2:57 PM

Ulrich,

I have two Z3801A and both of them act close to what you describe in the
first 24 hours of power up.  I believe its part of the disciplining
algorithm.  From what I read about the smart clock, it takes 5 days for
it to complete its initial learning cycle and then its continuously
refining.

Brian

Ulrich, I have two Z3801A and both of them act close to what you describe in the first 24 hours of power up. I believe its part of the disciplining algorithm. From what I read about the smart clock, it takes 5 days for it to complete its initial learning cycle and then its continuously refining. Brian
UB
Ulrich Bangert
Wed, Jun 17, 2009 5:44 PM

Brian,

thanks for your information!

algorithm.  From what I read about the smart clock, it takes
5 days for it to complete its initial learning cycle and then its

continuously

refining.

Is that from the Z3801 manual, which I have available, or do you have any
other in depth information source?

Best regards
Ulrich

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby
Gesendet: Mittwoch, 17. Juni 2009 16:58
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Ulrich,

I have two Z3801A and both of them act close to what you
describe in the
first 24 hours of power up.  I believe its part of the disciplining
algorithm.  From what I read about the smart clock, it takes
5 days for
it to complete its initial learning cycle and then its continuously
refining.

Brian


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Brian, thanks for your information! > algorithm. From what I read about the smart clock, it takes > 5 days for it to complete its initial learning cycle and then its continuously > refining. Is that from the Z3801 manual, which I have available, or do you have any other in depth information source? Best regards Ulrich > -----Ursprungliche Nachricht----- > Von: time-nuts-bounces@febo.com > [mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby > Gesendet: Mittwoch, 17. Juni 2009 16:58 > An: Discussion of precise time and frequency measurement > Betreff: Re: [time-nuts] Z3805 initial behaviour after power up > > > Ulrich, > > I have two Z3801A and both of them act close to what you > describe in the > first 24 hours of power up. I believe its part of the disciplining > algorithm. From what I read about the smart clock, it takes > 5 days for > it to complete its initial learning cycle and then its continuously > refining. > > Brian > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
MD
Magnus Danielson
Wed, Jun 17, 2009 7:31 PM

Hi Brian,

Brian Kirby skrev:

Ulrich,

I have two Z3801A and both of them act close to what you describe in the
first 24 hours of power up.  I believe its part of the disciplining
algorithm.  From what I read about the smart clock, it takes 5 days for
it to complete its initial learning cycle and then its continuously
refining.

For me it sounds more like a design-flaw than anything else. Using
several PLL bandwidths and switch between them is as such a good
approach, but the stepping between then needs to be done such that a
narrower bandwidth is only chosen when it the lock-in can be maintained.
Similarly, backing out of a narrow step to a wider step should also be
detected at suitable levels when it can't maintain track. The phase
detector gives hints about the ability to maintain track, as the phase
will deviate uncontrollably when loosing lock, but before that happens
it will deviate from near +/- 0 degrees, similarly, when within near +/-
0 degrees for sufficient time it is reasnoble that the next step (if not
too big) can maintain track. Recall that "sufficient time" changes with
the bandwidth of the PLL.

A third degree (PII^2 or PII^2D) PLL is better able to cope with drift
rate than a second degree PLL. A combined phase/frequency detection
approach (I.e. add the D term) adds quicker response to drift and
ability to keep tracking.

Another approach is to use a Kalman filter, where the Kalman gain is
adapted continuously. Kalman filters takes some careful thought, it's
not a magical wand to wave to make things better by magic. If done
properly, it will be able to fairly well track along and detect the
drift rate (takes a phase/frequency/drift model to handle) and update as
it stabilizes.

None of these approaches is rocket science to design anymore. HP/Agilent
surely could at the time of Z3801A and followers.

Cheers,
Magnus

Hi Brian, Brian Kirby skrev: > Ulrich, > > I have two Z3801A and both of them act close to what you describe in the > first 24 hours of power up. I believe its part of the disciplining > algorithm. From what I read about the smart clock, it takes 5 days for > it to complete its initial learning cycle and then its continuously > refining. For me it sounds more like a design-flaw than anything else. Using several PLL bandwidths and switch between them is as such a good approach, but the stepping between then needs to be done such that a narrower bandwidth is only chosen when it the lock-in can be maintained. Similarly, backing out of a narrow step to a wider step should also be detected at suitable levels when it can't maintain track. The phase detector gives hints about the ability to maintain track, as the phase will deviate uncontrollably when loosing lock, but before that happens it will deviate from near +/- 0 degrees, similarly, when within near +/- 0 degrees for sufficient time it is reasnoble that the next step (if not too big) can maintain track. Recall that "sufficient time" changes with the bandwidth of the PLL. A third degree (PII^2 or PII^2D) PLL is better able to cope with drift rate than a second degree PLL. A combined phase/frequency detection approach (I.e. add the D term) adds quicker response to drift and ability to keep tracking. Another approach is to use a Kalman filter, where the Kalman gain is adapted continuously. Kalman filters takes some careful thought, it's not a magical wand to wave to make things better by magic. If done properly, it will be able to fairly well track along and detect the drift rate (takes a phase/frequency/drift model to handle) and update as it stabilizes. None of these approaches is rocket science to design anymore. HP/Agilent surely could at the time of Z3801A and followers. Cheers, Magnus
DF
David Forbes
Wed, Jun 17, 2009 7:50 PM

Magnus Danielson wrote:

None of these approaches is rocket science to design anymore. HP/Agilent
surely could at the time of Z3801A and followers.

Yet the problem of an oscillator having a high drift rate after long-term
storage is not one that a development team is likely to be thinking about.

Nor is it a problem that can be easily examined, since it requires a bunch of
old units lying around the development lab. Old units don't exist when the
product is new, and even if they had traveled back in time to buy and age a few
units, after sufficient testing they would stop drifting and therefore be
useless for the test.

Nevertheless, 25 years later we expect the whiz-bang engineers at HP to have
thought of the problem and solved it.

--David Forbes, Tucson

Magnus Danielson wrote: > > None of these approaches is rocket science to design anymore. HP/Agilent > surely could at the time of Z3801A and followers. Yet the problem of an oscillator having a high drift rate after long-term storage is not one that a development team is likely to be thinking about. Nor is it a problem that can be easily examined, since it requires a bunch of old units lying around the development lab. Old units don't exist when the product is new, and even if they had traveled back in time to buy and age a few units, after sufficient testing they would stop drifting and therefore be useless for the test. Nevertheless, 25 years later we expect the whiz-bang engineers at HP to have thought of the problem and solved it. --David Forbes, Tucson
FL
Francesco Ledda
Wed, Jun 17, 2009 7:53 PM

A third order PPL has better sideband suppression.  The tracking capability
depends upon the phase range of the phase detector.  If 1:1 phase comparison
and a XOR phase detector is used, the range is 180 degrees. Dividers on the
feedback portion of the PLL increase the phase range of the phase detector.
For example, if the expected jitter is 10UI, a divider larger than 10 must
be used, to maintain lock under all conditions. A monotonically decreasing
phase delta, on the phase detector, still means that the PLL is locked.

The phase/frequency detector, avoids the need for a frequency aquisition
aid.  Once a phase reversal is detected by the flip flops in the phase/freq
detector, it goes back to phase detector mode.

My experience is that most fancy syncrhonizer for telecom application
(Stratum, LORAN and GPS) use start-stop phase detector with averager in
front of the AC-DC gain circuits.  Changing the closed loop bandwidth of the
PLL on the fly is not easy, due to secondary effects (done that many times).

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On
Behalf Of Magnus Danielson
Sent: Wednesday, June 17, 2009 2:31 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Z3805 initial behaviour after power up

Hi Brian,

Brian Kirby skrev:

Ulrich,

I have two Z3801A and both of them act close to what you describe in the
first 24 hours of power up.  I believe its part of the disciplining
algorithm.  From what I read about the smart clock, it takes 5 days for
it to complete its initial learning cycle and then its continuously
refining.

For me it sounds more like a design-flaw than anything else. Using
several PLL bandwidths and switch between them is as such a good
approach, but the stepping between then needs to be done such that a
narrower bandwidth is only chosen when it the lock-in can be maintained.
Similarly, backing out of a narrow step to a wider step should also be
detected at suitable levels when it can't maintain track. The phase
detector gives hints about the ability to maintain track, as the phase
will deviate uncontrollably when loosing lock, but before that happens
it will deviate from near +/- 0 degrees, similarly, when within near +/-
0 degrees for sufficient time it is reasnoble that the next step (if not
too big) can maintain track. Recall that "sufficient time" changes with
the bandwidth of the PLL.

A third degree (PII^2 or PII^2D) PLL is better able to cope with drift
rate than a second degree PLL. A combined phase/frequency detection
approach (I.e. add the D term) adds quicker response to drift and
ability to keep tracking.

Another approach is to use a Kalman filter, where the Kalman gain is
adapted continuously. Kalman filters takes some careful thought, it's
not a magical wand to wave to make things better by magic. If done
properly, it will be able to fairly well track along and detect the
drift rate (takes a phase/frequency/drift model to handle) and update as
it stabilizes.

None of these approaches is rocket science to design anymore. HP/Agilent
surely could at the time of Z3801A and followers.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

A third order PPL has better sideband suppression. The tracking capability depends upon the phase range of the phase detector. If 1:1 phase comparison and a XOR phase detector is used, the range is 180 degrees. Dividers on the feedback portion of the PLL increase the phase range of the phase detector. For example, if the expected jitter is 10UI, a divider larger than 10 must be used, to maintain lock under all conditions. A monotonically decreasing phase delta, on the phase detector, still means that the PLL is locked. The phase/frequency detector, avoids the need for a frequency aquisition aid. Once a phase reversal is detected by the flip flops in the phase/freq detector, it goes back to phase detector mode. My experience is that most fancy syncrhonizer for telecom application (Stratum, LORAN and GPS) use start-stop phase detector with averager in front of the AC-DC gain circuits. Changing the closed loop bandwidth of the PLL on the fly is not easy, due to secondary effects (done that many times). -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On Behalf Of Magnus Danielson Sent: Wednesday, June 17, 2009 2:31 PM To: Discussion of precise time and frequency measurement Subject: Re: [time-nuts] Z3805 initial behaviour after power up Hi Brian, Brian Kirby skrev: > Ulrich, > > I have two Z3801A and both of them act close to what you describe in the > first 24 hours of power up. I believe its part of the disciplining > algorithm. From what I read about the smart clock, it takes 5 days for > it to complete its initial learning cycle and then its continuously > refining. For me it sounds more like a design-flaw than anything else. Using several PLL bandwidths and switch between them is as such a good approach, but the stepping between then needs to be done such that a narrower bandwidth is only chosen when it the lock-in can be maintained. Similarly, backing out of a narrow step to a wider step should also be detected at suitable levels when it can't maintain track. The phase detector gives hints about the ability to maintain track, as the phase will deviate uncontrollably when loosing lock, but before that happens it will deviate from near +/- 0 degrees, similarly, when within near +/- 0 degrees for sufficient time it is reasnoble that the next step (if not too big) can maintain track. Recall that "sufficient time" changes with the bandwidth of the PLL. A third degree (PII^2 or PII^2D) PLL is better able to cope with drift rate than a second degree PLL. A combined phase/frequency detection approach (I.e. add the D term) adds quicker response to drift and ability to keep tracking. Another approach is to use a Kalman filter, where the Kalman gain is adapted continuously. Kalman filters takes some careful thought, it's not a magical wand to wave to make things better by magic. If done properly, it will be able to fairly well track along and detect the drift rate (takes a phase/frequency/drift model to handle) and update as it stabilizes. None of these approaches is rocket science to design anymore. HP/Agilent surely could at the time of Z3801A and followers. Cheers, Magnus _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
MD
Magnus Danielson
Wed, Jun 17, 2009 9:53 PM

David Forbes skrev:

Magnus Danielson wrote:

None of these approaches is rocket science to design anymore.
HP/Agilent surely could at the time of Z3801A and followers.

Yet the problem of an oscillator having a high drift rate after
long-term storage is not one that a development team is likely to be
thinking about.

Nor is it a problem that can be easily examined, since it requires a
bunch of old units lying around the development lab. Old units don't
exist when the product is new, and even if they had traveled back in
time to buy and age a few units, after sufficient testing they would
stop drifting and therefore be useless for the test.

Nevertheless, 25 years later we expect the whiz-bang engineers at HP to
have thought of the problem and solved it.

Starting any oscillator from cold is pretty upsetting and takes time to
drift in. But sure, it is kind of not so probable that they would
encounter it.

But still, crystal oscillators is individuals and there are other forms
of abuse (such as chock) that can occur in an oscillators life... so you
would expect that some anticipation of a "less ideal case" might have
passed their minds. Ah well, maybe that is me... I hate seeing returns
from the field only due to fairly easy mistakes in the design. Not that
it doesn't happens, but it's a learning process to improve on design margin.

Cheers,
Magnus

David Forbes skrev: > Magnus Danielson wrote: >> >> None of these approaches is rocket science to design anymore. >> HP/Agilent surely could at the time of Z3801A and followers. > > Yet the problem of an oscillator having a high drift rate after > long-term storage is not one that a development team is likely to be > thinking about. > > Nor is it a problem that can be easily examined, since it requires a > bunch of old units lying around the development lab. Old units don't > exist when the product is new, and even if they had traveled back in > time to buy and age a few units, after sufficient testing they would > stop drifting and therefore be useless for the test. > > Nevertheless, 25 years later we expect the whiz-bang engineers at HP to > have thought of the problem and solved it. Starting any oscillator from cold is pretty upsetting and takes time to drift in. But sure, it is kind of not so probable that they would encounter it. But still, crystal oscillators is individuals and there are other forms of abuse (such as chock) that can occur in an oscillators life... so you would expect that some anticipation of a "less ideal case" might have passed their minds. Ah well, maybe that is me... I hate seeing returns from the field only due to fairly easy mistakes in the design. Not that it doesn't happens, but it's a learning process to improve on design margin. Cheers, Magnus
MD
Magnus Danielson
Wed, Jun 17, 2009 10:09 PM

Hi Francesco,

Francesco Ledda skrev:

A third order PPL has better sideband suppression.  The tracking capability
depends upon the phase range of the phase detector.  If 1:1 phase comparison
and a XOR phase detector is used, the range is 180 degrees. Dividers on the
feedback portion of the PLL increase the phase range of the phase detector.
For example, if the expected jitter is 10UI, a divider larger than 10 must
be used, to maintain lock under all conditions. A monotonically decreasing
phase delta, on the phase detector, still means that the PLL is locked.

True, but regardless of how large you have made your phase detectors
range, if you have a loop unable to track to the dynamics you put into
it, you eventually reach the wrapping or limiting point. By back out
into quicker response, the PLL is able to quicker reduce the phase error
and track it in, as it quicker adjust its state to match the difference
between the source and the oscillator.

The phase/frequency detector, avoids the need for a frequency aquisition
aid.  Once a phase reversal is detected by the flip flops in the phase/freq
detector, it goes back to phase detector mode.

Sure, but it does not totally remove the benefits of shifting PLL
bandwidth according to need. I'm all for phase/frequency detectors...
and have designed several.

My experience is that most fancy syncrhonizer for telecom application
(Stratum, LORAN and GPS) use start-stop phase detector with averager in
front of the AC-DC gain circuits.  Changing the closed loop bandwidth of the
PLL on the fly is not easy, due to secondary effects (done that many times).

Such a detector isn't particularly useful for most GPSDOs as they get
have 1 Hz comparator frequency with the PPS pulse. The analog
time-constants would be a terrible mess to design. For GPSDOs using PPS
updates a digital loop filter can be assumed (unless you use say the
Jupiter receiver and it's 10 kHz output, see performance on TvB pages!).

Being able to scale loop parameters without having to re-scale state is
indeed an issue, but there exists topologies that can achieve this
without the need for rescaling. These can be realized both in the
analogue world using say OTAs and or through multipliers at the right
positions when done in the digital world. It is used with great success
in the GPS code and carrier tracking channels. You can look it up in the
Kaplan GPS book for instance... which only summarize what's found in
other places.

Cheers,
Magnus

Hi Francesco, Francesco Ledda skrev: > A third order PPL has better sideband suppression. The tracking capability > depends upon the phase range of the phase detector. If 1:1 phase comparison > and a XOR phase detector is used, the range is 180 degrees. Dividers on the > feedback portion of the PLL increase the phase range of the phase detector. > For example, if the expected jitter is 10UI, a divider larger than 10 must > be used, to maintain lock under all conditions. A monotonically decreasing > phase delta, on the phase detector, still means that the PLL is locked. True, but regardless of how large you have made your phase detectors range, if you have a loop unable to track to the dynamics you put into it, you eventually reach the wrapping or limiting point. By back out into quicker response, the PLL is able to quicker reduce the phase error and track it in, as it quicker adjust its state to match the difference between the source and the oscillator. > The phase/frequency detector, avoids the need for a frequency aquisition > aid. Once a phase reversal is detected by the flip flops in the phase/freq > detector, it goes back to phase detector mode. Sure, but it does not totally remove the benefits of shifting PLL bandwidth according to need. I'm all for phase/frequency detectors... and have designed several. > My experience is that most fancy syncrhonizer for telecom application > (Stratum, LORAN and GPS) use start-stop phase detector with averager in > front of the AC-DC gain circuits. Changing the closed loop bandwidth of the > PLL on the fly is not easy, due to secondary effects (done that many times). Such a detector isn't particularly useful for most GPSDOs as they get have 1 Hz comparator frequency with the PPS pulse. The analog time-constants would be a terrible mess to design. For GPSDOs using PPS updates a digital loop filter can be assumed (unless you use say the Jupiter receiver and it's 10 kHz output, see performance on TvB pages!). Being able to scale loop parameters without having to re-scale state is indeed an issue, but there exists topologies that can achieve this without the need for rescaling. These can be realized both in the analogue world using say OTAs and or through multipliers at the right positions when done in the digital world. It is used with great success in the GPS code and carrier tracking channels. You can look it up in the Kaplan GPS book for instance... which only summarize what's found in other places. Cheers, Magnus
BK
Brian Kirby
Wed, Jun 17, 2009 11:50 PM

They do not talk about driving the frequency off, but they warn to keep
the receiver locked the first 24 hours, so it can determine how the
oscillator ages.  Its in the section on "holdover" (page 52 of the PDF ,
page 3-8 of the user guide).

You may also want to search for HP An1279, it also goes over HP
smartclock operation, and look for a paper titled "the Global
Positioning System and HP SmartClock" by John A. Kusters.

If you have a broadband connection, I can email you each one, if that
will help.

Brian

Ulrich Bangert wrote:

Brian,

thanks for your information!

algorithm.  From what I read about the smart clock, it takes
5 days for it to complete its initial learning cycle and then its

continuously

refining.

Is that from the Z3801 manual, which I have available, or do you have any
other in depth information source?

Best regards
Ulrich

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby
Gesendet: Mittwoch, 17. Juni 2009 16:58
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Ulrich,

I have two Z3801A and both of them act close to what you
describe in the
first 24 hours of power up.  I believe its part of the disciplining
algorithm.  From what I read about the smart clock, it takes
5 days for
it to complete its initial learning cycle and then its continuously
refining.

Brian


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

They do not talk about driving the frequency off, but they warn to keep the receiver locked the first 24 hours, so it can determine how the oscillator ages. Its in the section on "holdover" (page 52 of the PDF , page 3-8 of the user guide). You may also want to search for HP An1279, it also goes over HP smartclock operation, and look for a paper titled "the Global Positioning System and HP SmartClock" by John A. Kusters. If you have a broadband connection, I can email you each one, if that will help. Brian Ulrich Bangert wrote: > Brian, > > thanks for your information! > > >> algorithm. From what I read about the smart clock, it takes >> 5 days for it to complete its initial learning cycle and then its >> > continuously > >> refining. >> > > Is that from the Z3801 manual, which I have available, or do you have any > other in depth information source? > > Best regards > Ulrich > > >> -----Ursprungliche Nachricht----- >> Von: time-nuts-bounces@febo.com >> [mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby >> Gesendet: Mittwoch, 17. Juni 2009 16:58 >> An: Discussion of precise time and frequency measurement >> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up >> >> >> Ulrich, >> >> I have two Z3801A and both of them act close to what you >> describe in the >> first 24 hours of power up. I believe its part of the disciplining >> algorithm. From what I read about the smart clock, it takes >> 5 days for >> it to complete its initial learning cycle and then its continuously >> refining. >> >> Brian >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > >
UB
Ulrich Bangert
Thu, Jun 18, 2009 6:28 AM

Brian,

You may also want to search for HP An1279, it also goes over HP
smartclock operation, and look for a paper titled "the Global
Positioning System and HP SmartClock" by John A. Kusters.

In the meantime I have not only found these but also "Smart Clock: A New
Time" by David Allan et al which shows that "Smart Clock" is originally a
NIST invention & patent and explains very well how it works. Perhaps what I
and You have seen is a direct consequence of applying the Smart Clock
algorithm.

Best regards
Ulrich

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby
Gesendet: Donnerstag, 18. Juni 2009 01:51
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

They do not talk about driving the frequency off, but they
warn to keep
the receiver locked the first 24 hours, so it can determine how the
oscillator ages.  Its in the section on "holdover" (page 52
of the PDF ,
page 3-8 of the user guide).

You may also want to search for HP An1279, it also goes over HP
smartclock operation, and look for a paper titled "the Global
Positioning System and HP SmartClock" by John A. Kusters.

If you have a broadband connection, I can email you each one, if that
will help.

Brian

Ulrich Bangert wrote:

Brian,

thanks for your information!

algorithm.  From what I read about the smart clock, it takes
5 days for it to complete its initial learning cycle and then its

continuously

refining.

Is that from the Z3801 manual, which I have available, or

do you have

any other in depth information source?

Best regards
Ulrich

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby
Gesendet: Mittwoch, 17. Juni 2009 16:58
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Ulrich,

I have two Z3801A and both of them act close to what you
describe in the
first 24 hours of power up.  I believe its part of the

disciplining

algorithm.  From what I read about the smart clock, it takes
5 days for
it to complete its initial learning cycle and then its

continuously

refining.

Brian


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Brian, > You may also want to search for HP An1279, it also goes over HP > smartclock operation, and look for a paper titled "the Global > Positioning System and HP SmartClock" by John A. Kusters. In the meantime I have not only found these but also "Smart Clock: A New Time" by David Allan et al which shows that "Smart Clock" is originally a NIST invention & patent and explains very well how it works. Perhaps what I and You have seen is a direct consequence of applying the Smart Clock algorithm. Best regards Ulrich > -----Ursprungliche Nachricht----- > Von: time-nuts-bounces@febo.com > [mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby > Gesendet: Donnerstag, 18. Juni 2009 01:51 > An: Discussion of precise time and frequency measurement > Betreff: Re: [time-nuts] Z3805 initial behaviour after power up > > > They do not talk about driving the frequency off, but they > warn to keep > the receiver locked the first 24 hours, so it can determine how the > oscillator ages. Its in the section on "holdover" (page 52 > of the PDF , > page 3-8 of the user guide). > > You may also want to search for HP An1279, it also goes over HP > smartclock operation, and look for a paper titled "the Global > Positioning System and HP SmartClock" by John A. Kusters. > > If you have a broadband connection, I can email you each one, if that > will help. > > Brian > > Ulrich Bangert wrote: > > Brian, > > > > thanks for your information! > > > > > >> algorithm. From what I read about the smart clock, it takes > >> 5 days for it to complete its initial learning cycle and then its > >> > > continuously > > > >> refining. > >> > > > > Is that from the Z3801 manual, which I have available, or > do you have > > any other in depth information source? > > > > Best regards > > Ulrich > > > > > >> -----Ursprungliche Nachricht----- > >> Von: time-nuts-bounces@febo.com > >> [mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby > >> Gesendet: Mittwoch, 17. Juni 2009 16:58 > >> An: Discussion of precise time and frequency measurement > >> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up > >> > >> > >> Ulrich, > >> > >> I have two Z3801A and both of them act close to what you > >> describe in the > >> first 24 hours of power up. I believe its part of the > disciplining > >> algorithm. From what I read about the smart clock, it takes > >> 5 days for > >> it to complete its initial learning cycle and then its > continuously > >> refining. > >> > >> Brian > >> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@febo.com > >> To unsubscribe, go to > >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> and follow the instructions there. > >> > > > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
UB
Ulrich Bangert
Thu, Jun 18, 2009 8:04 AM

Gents,

one of the papers suggested by Brian says:

SmartClock monitors the frequency control variable of the internal
oscillator while it is locked to the external reference. This gives a
measure of the frequency difference between the internal oscillator, if it
is free-running, and the external reference over time. The resulting
measurements include the effects of random noise in the oscillator, the
measurement circuitry, and any noise in the external reference as well as
any aging and environmental effects in the oscillator. From this
information, SmartClock makes a continuous prediction of clock error over
time.

The big question is: If the EFC signal includes this all information, how
does the algo manage to extract the individual informations? After having a
look at the PPS TI and the EFC of my Z3805 I am beginning to get a clue of
it (the quirks on the EFC signal heve been removed):

The SmartClock algo seems to set the loop time constants to a large value in
the beginning. In other words: It measures the overall oscillator frequency
influencing effects with a lowpass filter applied that has a very low cutoff
frequency. Only effects which's frequency are sufficient lower then the
cutoff pass the filter. Seems as if the SmartClock uses this filter setting
to exclude any noise related effect and any environmental effect and make
the measurement see ONLY the oscillator's aging.

Once it has learned enough about the aging process it will be able to
predict aging. Having reached this state it will be able to apply his
knowledge about aging to further TI measurements and subtract the "aging
part" of it, leaving mostly the environmental (which is mostly temperature)
part.

Having a model for the aging the Algo can now model the temperature
dependence. However, a precondition for this were that the measurement sees
the temperature effects and that the ambient temperature is measured
independently. For that reason I expect the Algo will reduce its loop time
constant within the next time to make the temperature effect get through the
lowpass. Any bets on that?

Best regards
Ulrich Bangert

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Ulrich Bangert
Gesendet: Donnerstag, 18. Juni 2009 08:29
An: 'Discussion of precise time and frequency measurement'
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Brian,

You may also want to search for HP An1279, it also goes over HP
smartclock operation, and look for a paper titled "the Global
Positioning System and HP SmartClock" by John A. Kusters.

In the meantime I have not only found these but also "Smart
Clock: A New Time" by David Allan et al which shows that
"Smart Clock" is originally a NIST invention & patent and
explains very well how it works. Perhaps what I and You have
seen is a direct consequence of applying the Smart Clock algorithm.

Best regards
Ulrich

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby
Gesendet: Donnerstag, 18. Juni 2009 01:51
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

They do not talk about driving the frequency off, but they
warn to keep
the receiver locked the first 24 hours, so it can determine how the
oscillator ages.  Its in the section on "holdover" (page 52
of the PDF ,
page 3-8 of the user guide).

You may also want to search for HP An1279, it also goes over HP
smartclock operation, and look for a paper titled "the Global
Positioning System and HP SmartClock" by John A. Kusters.

If you have a broadband connection, I can email you each

one, if that

will help.

Brian

Ulrich Bangert wrote:

Brian,

thanks for your information!

algorithm.  From what I read about the smart clock, it

takes 5 days

for it to complete its initial learning cycle and then its

continuously

refining.

Is that from the Z3801 manual, which I have available, or

do you have

any other in depth information source?

Best regards
Ulrich

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com

Im Auftrag von Brian Kirby
Gesendet: Mittwoch, 17. Juni 2009 16:58
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Ulrich,

I have two Z3801A and both of them act close to what you

describe

in the first 24 hours of power up.  I believe its part of the

disciplining

algorithm.  From what I read about the smart clock, it takes
5 days for
it to complete its initial learning cycle and then its

continuously

refining.

Brian


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Gents, one of the papers suggested by Brian says: ------------------------------------------- SmartClock monitors the frequency control variable of the internal oscillator while it is locked to the external reference. This gives a measure of the frequency difference between the internal oscillator, if it is free-running, and the external reference over time. The resulting measurements include the effects of random noise in the oscillator, the measurement circuitry, and any noise in the external reference as well as any aging and environmental effects in the oscillator. From this information, SmartClock makes a continuous prediction of clock error over time. ------------------------------------------- The big question is: If the EFC signal includes this all information, how does the algo manage to extract the individual informations? After having a look at the PPS TI and the EFC of my Z3805 I am beginning to get a clue of it (the quirks on the EFC signal heve been removed): The SmartClock algo seems to set the loop time constants to a large value in the beginning. In other words: It measures the overall oscillator frequency influencing effects with a lowpass filter applied that has a very low cutoff frequency. Only effects which's frequency are sufficient lower then the cutoff pass the filter. Seems as if the SmartClock uses this filter setting to exclude any noise related effect and any environmental effect and make the measurement see ONLY the oscillator's aging. Once it has learned enough about the aging process it will be able to predict aging. Having reached this state it will be able to apply his knowledge about aging to further TI measurements and subtract the "aging part" of it, leaving mostly the environmental (which is mostly temperature) part. Having a model for the aging the Algo can now model the temperature dependence. However, a precondition for this were that the measurement sees the temperature effects and that the ambient temperature is measured independently. For that reason I expect the Algo will reduce its loop time constant within the next time to make the temperature effect get through the lowpass. Any bets on that? Best regards Ulrich Bangert > -----Ursprungliche Nachricht----- > Von: time-nuts-bounces@febo.com > [mailto:time-nuts-bounces@febo.com] Im Auftrag von Ulrich Bangert > Gesendet: Donnerstag, 18. Juni 2009 08:29 > An: 'Discussion of precise time and frequency measurement' > Betreff: Re: [time-nuts] Z3805 initial behaviour after power up > > > Brian, > > > You may also want to search for HP An1279, it also goes over HP > > smartclock operation, and look for a paper titled "the Global > > Positioning System and HP SmartClock" by John A. Kusters. > > In the meantime I have not only found these but also "Smart > Clock: A New Time" by David Allan et al which shows that > "Smart Clock" is originally a NIST invention & patent and > explains very well how it works. Perhaps what I and You have > seen is a direct consequence of applying the Smart Clock algorithm. > > Best regards > Ulrich > > > > -----Ursprungliche Nachricht----- > > Von: time-nuts-bounces@febo.com > > [mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby > > Gesendet: Donnerstag, 18. Juni 2009 01:51 > > An: Discussion of precise time and frequency measurement > > Betreff: Re: [time-nuts] Z3805 initial behaviour after power up > > > > > > They do not talk about driving the frequency off, but they > > warn to keep > > the receiver locked the first 24 hours, so it can determine how the > > oscillator ages. Its in the section on "holdover" (page 52 > > of the PDF , > > page 3-8 of the user guide). > > > > You may also want to search for HP An1279, it also goes over HP > > smartclock operation, and look for a paper titled "the Global > > Positioning System and HP SmartClock" by John A. Kusters. > > > > If you have a broadband connection, I can email you each > one, if that > > will help. > > > > Brian > > > > Ulrich Bangert wrote: > > > Brian, > > > > > > thanks for your information! > > > > > > > > >> algorithm. From what I read about the smart clock, it > takes 5 days > > >> for it to complete its initial learning cycle and then its > > >> > > > continuously > > > > > >> refining. > > >> > > > > > > Is that from the Z3801 manual, which I have available, or > > do you have > > > any other in depth information source? > > > > > > Best regards > > > Ulrich > > > > > > > > >> -----Ursprungliche Nachricht----- > > >> Von: time-nuts-bounces@febo.com > [mailto:time-nuts-bounces@febo.com] > > >> Im Auftrag von Brian Kirby > > >> Gesendet: Mittwoch, 17. Juni 2009 16:58 > > >> An: Discussion of precise time and frequency measurement > > >> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up > > >> > > >> > > >> Ulrich, > > >> > > >> I have two Z3801A and both of them act close to what you > describe > > >> in the first 24 hours of power up. I believe its part of the > > disciplining > > >> algorithm. From what I read about the smart clock, it takes > > >> 5 days for > > >> it to complete its initial learning cycle and then its > > continuously > > >> refining. > > >> > > >> Brian > > >> > > >> _______________________________________________ > > >> time-nuts mailing list -- time-nuts@febo.com > > >> To unsubscribe, go to > > >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > >> and follow the instructions there. > > >> > > > > > > > > > _______________________________________________ > > > time-nuts mailing list -- time-nuts@febo.com > > > To unsubscribe, go to > > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > > and follow the instructions there. > > > > > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
FL
Francesco Ledda
Thu, Jun 18, 2009 1:06 PM

Aging cannot be predicted!  If it could be predicetd, there would be no
aging.

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On
Behalf Of Ulrich Bangert
Sent: Thursday, June 18, 2009 3:04 AM
To: 'Discussion of precise time and frequency measurement'
Subject: [time-nuts] Z3805 initial behaviour after power up revised

Gents,

one of the papers suggested by Brian says:

SmartClock monitors the frequency control variable of the internal
oscillator while it is locked to the external reference. This gives a
measure of the frequency difference between the internal oscillator, if it
is free-running, and the external reference over time. The resulting
measurements include the effects of random noise in the oscillator, the
measurement circuitry, and any noise in the external reference as well as
any aging and environmental effects in the oscillator. From this
information, SmartClock makes a continuous prediction of clock error over
time.

The big question is: If the EFC signal includes this all information, how
does the algo manage to extract the individual informations? After having a
look at the PPS TI and the EFC of my Z3805 I am beginning to get a clue of
it (the quirks on the EFC signal heve been removed):

The SmartClock algo seems to set the loop time constants to a large value in
the beginning. In other words: It measures the overall oscillator frequency
influencing effects with a lowpass filter applied that has a very low cutoff
frequency. Only effects which's frequency are sufficient lower then the
cutoff pass the filter. Seems as if the SmartClock uses this filter setting
to exclude any noise related effect and any environmental effect and make
the measurement see ONLY the oscillator's aging.

Once it has learned enough about the aging process it will be able to
predict aging. Having reached this state it will be able to apply his
knowledge about aging to further TI measurements and subtract the "aging
part" of it, leaving mostly the environmental (which is mostly temperature)
part.

Having a model for the aging the Algo can now model the temperature
dependence. However, a precondition for this were that the measurement sees
the temperature effects and that the ambient temperature is measured
independently. For that reason I expect the Algo will reduce its loop time
constant within the next time to make the temperature effect get through the
lowpass. Any bets on that?

Best regards
Ulrich Bangert

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Ulrich Bangert
Gesendet: Donnerstag, 18. Juni 2009 08:29
An: 'Discussion of precise time and frequency measurement'
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Brian,

You may also want to search for HP An1279, it also goes over HP
smartclock operation, and look for a paper titled "the Global
Positioning System and HP SmartClock" by John A. Kusters.

In the meantime I have not only found these but also "Smart
Clock: A New Time" by David Allan et al which shows that
"Smart Clock" is originally a NIST invention & patent and
explains very well how it works. Perhaps what I and You have
seen is a direct consequence of applying the Smart Clock algorithm.

Best regards
Ulrich

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby
Gesendet: Donnerstag, 18. Juni 2009 01:51
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

They do not talk about driving the frequency off, but they
warn to keep
the receiver locked the first 24 hours, so it can determine how the
oscillator ages.  Its in the section on "holdover" (page 52
of the PDF ,
page 3-8 of the user guide).

You may also want to search for HP An1279, it also goes over HP
smartclock operation, and look for a paper titled "the Global
Positioning System and HP SmartClock" by John A. Kusters.

If you have a broadband connection, I can email you each

one, if that

will help.

Brian

Ulrich Bangert wrote:

Brian,

thanks for your information!

algorithm.  From what I read about the smart clock, it

takes 5 days

for it to complete its initial learning cycle and then its

continuously

refining.

Is that from the Z3801 manual, which I have available, or

do you have

any other in depth information source?

Best regards
Ulrich

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com

Im Auftrag von Brian Kirby
Gesendet: Mittwoch, 17. Juni 2009 16:58
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Ulrich,

I have two Z3801A and both of them act close to what you

describe

in the first 24 hours of power up.  I believe its part of the

disciplining

algorithm.  From what I read about the smart clock, it takes
5 days for
it to complete its initial learning cycle and then its

continuously

refining.

Brian


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Aging cannot be predicted! If it could be predicetd, there would be no aging. -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On Behalf Of Ulrich Bangert Sent: Thursday, June 18, 2009 3:04 AM To: 'Discussion of precise time and frequency measurement' Subject: [time-nuts] Z3805 initial behaviour after power up revised Gents, one of the papers suggested by Brian says: ------------------------------------------- SmartClock monitors the frequency control variable of the internal oscillator while it is locked to the external reference. This gives a measure of the frequency difference between the internal oscillator, if it is free-running, and the external reference over time. The resulting measurements include the effects of random noise in the oscillator, the measurement circuitry, and any noise in the external reference as well as any aging and environmental effects in the oscillator. From this information, SmartClock makes a continuous prediction of clock error over time. ------------------------------------------- The big question is: If the EFC signal includes this all information, how does the algo manage to extract the individual informations? After having a look at the PPS TI and the EFC of my Z3805 I am beginning to get a clue of it (the quirks on the EFC signal heve been removed): The SmartClock algo seems to set the loop time constants to a large value in the beginning. In other words: It measures the overall oscillator frequency influencing effects with a lowpass filter applied that has a very low cutoff frequency. Only effects which's frequency are sufficient lower then the cutoff pass the filter. Seems as if the SmartClock uses this filter setting to exclude any noise related effect and any environmental effect and make the measurement see ONLY the oscillator's aging. Once it has learned enough about the aging process it will be able to predict aging. Having reached this state it will be able to apply his knowledge about aging to further TI measurements and subtract the "aging part" of it, leaving mostly the environmental (which is mostly temperature) part. Having a model for the aging the Algo can now model the temperature dependence. However, a precondition for this were that the measurement sees the temperature effects and that the ambient temperature is measured independently. For that reason I expect the Algo will reduce its loop time constant within the next time to make the temperature effect get through the lowpass. Any bets on that? Best regards Ulrich Bangert > -----Ursprungliche Nachricht----- > Von: time-nuts-bounces@febo.com > [mailto:time-nuts-bounces@febo.com] Im Auftrag von Ulrich Bangert > Gesendet: Donnerstag, 18. Juni 2009 08:29 > An: 'Discussion of precise time and frequency measurement' > Betreff: Re: [time-nuts] Z3805 initial behaviour after power up > > > Brian, > > > You may also want to search for HP An1279, it also goes over HP > > smartclock operation, and look for a paper titled "the Global > > Positioning System and HP SmartClock" by John A. Kusters. > > In the meantime I have not only found these but also "Smart > Clock: A New Time" by David Allan et al which shows that > "Smart Clock" is originally a NIST invention & patent and > explains very well how it works. Perhaps what I and You have > seen is a direct consequence of applying the Smart Clock algorithm. > > Best regards > Ulrich > > > > -----Ursprungliche Nachricht----- > > Von: time-nuts-bounces@febo.com > > [mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby > > Gesendet: Donnerstag, 18. Juni 2009 01:51 > > An: Discussion of precise time and frequency measurement > > Betreff: Re: [time-nuts] Z3805 initial behaviour after power up > > > > > > They do not talk about driving the frequency off, but they > > warn to keep > > the receiver locked the first 24 hours, so it can determine how the > > oscillator ages. Its in the section on "holdover" (page 52 > > of the PDF , > > page 3-8 of the user guide). > > > > You may also want to search for HP An1279, it also goes over HP > > smartclock operation, and look for a paper titled "the Global > > Positioning System and HP SmartClock" by John A. Kusters. > > > > If you have a broadband connection, I can email you each > one, if that > > will help. > > > > Brian > > > > Ulrich Bangert wrote: > > > Brian, > > > > > > thanks for your information! > > > > > > > > >> algorithm. From what I read about the smart clock, it > takes 5 days > > >> for it to complete its initial learning cycle and then its > > >> > > > continuously > > > > > >> refining. > > >> > > > > > > Is that from the Z3801 manual, which I have available, or > > do you have > > > any other in depth information source? > > > > > > Best regards > > > Ulrich > > > > > > > > >> -----Ursprungliche Nachricht----- > > >> Von: time-nuts-bounces@febo.com > [mailto:time-nuts-bounces@febo.com] > > >> Im Auftrag von Brian Kirby > > >> Gesendet: Mittwoch, 17. Juni 2009 16:58 > > >> An: Discussion of precise time and frequency measurement > > >> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up > > >> > > >> > > >> Ulrich, > > >> > > >> I have two Z3801A and both of them act close to what you > describe > > >> in the first 24 hours of power up. I believe its part of the > > disciplining > > >> algorithm. From what I read about the smart clock, it takes > > >> 5 days for > > >> it to complete its initial learning cycle and then its > > continuously > > >> refining. > > >> > > >> Brian > > >> > > >> _______________________________________________ > > >> time-nuts mailing list -- time-nuts@febo.com > > >> To unsubscribe, go to > > >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > >> and follow the instructions there. > > >> > > > > > > > > > _______________________________________________ > > > time-nuts mailing list -- time-nuts@febo.com > > > To unsubscribe, go to > > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > > and follow the instructions there. > > > > > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
MW
M. Warner Losh
Fri, Jun 19, 2009 3:28 AM

In message: PAEJKNPPIGMNDLMJLENMGEFJHNAA.frledda@verizon.net
"Francesco Ledda" frledda@verizon.net writes:
: Aging cannot be predicted!  If it could be predicetd, there would be no
: aging.

Actually, if aging could be predicted perfectly, there'd be perfect
holdover (which I think is saying the same thing).  Aging can be
predicted imperfectly in the short term, but the amount of imperfect
grows with the time you're without data...

Warner

:
: -----Original Message-----
: From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On
: Behalf Of Ulrich Bangert
: Sent: Thursday, June 18, 2009 3:04 AM
: To: 'Discussion of precise time and frequency measurement'
: Subject: [time-nuts] Z3805 initial behaviour after power up revised
:
:
: Gents,
:
: one of the papers suggested by Brian says:
: -------------------------------------------
: SmartClock monitors the frequency control variable of the internal
: oscillator while it is locked to the external reference. This gives a
: measure of the frequency difference between the internal oscillator, if it
: is free-running, and the external reference over time. The resulting
: measurements include the effects of random noise in the oscillator, the
: measurement circuitry, and any noise in the external reference as well as
: any aging and environmental effects in the oscillator. From this
: information, SmartClock makes a continuous prediction of clock error over
: time.
: -------------------------------------------
:
: The big question is: If the EFC signal includes this all information, how
: does the algo manage to extract the individual informations? After having a
: look at the PPS TI and the EFC of my Z3805 I am beginning to get a clue of
: it (the quirks on the EFC signal heve been removed):
:
: The SmartClock algo seems to set the loop time constants to a large value in
: the beginning. In other words: It measures the overall oscillator frequency
: influencing effects with a lowpass filter applied that has a very low cutoff
: frequency. Only effects which's frequency are sufficient lower then the
: cutoff pass the filter. Seems as if the SmartClock uses this filter setting
: to exclude any noise related effect and any environmental effect and make
: the measurement see ONLY the oscillator's aging.
:
: Once it has learned enough about the aging process it will be able to
: predict aging. Having reached this state it will be able to apply his
: knowledge about aging to further TI measurements and subtract the "aging
: part" of it, leaving mostly the environmental (which is mostly temperature)
: part.
:
: Having a model for the aging the Algo can now model the temperature
: dependence. However, a precondition for this were that the measurement sees
: the temperature effects and that the ambient temperature is measured
: independently. For that reason I expect the Algo will reduce its loop time
: constant within the next time to make the temperature effect get through the
: lowpass. Any bets on that?
:
: Best regards
: Ulrich Bangert
:
: > -----Ursprungliche Nachricht-----
: > Von: time-nuts-bounces@febo.com
: > [mailto:time-nuts-bounces@febo.com] Im Auftrag von Ulrich Bangert
: > Gesendet: Donnerstag, 18. Juni 2009 08:29
: > An: 'Discussion of precise time and frequency measurement'
: > Betreff: Re: [time-nuts] Z3805 initial behaviour after power up
: >
: >
: > Brian,
: >
: > > You may also want to search for HP An1279, it also goes over HP
: > > smartclock operation, and look for a paper titled "the Global
: > > Positioning System and HP SmartClock" by John A. Kusters.
: >
: > In the meantime I have not only found these but also "Smart
: > Clock: A New Time" by David Allan et al which shows that
: > "Smart Clock" is originally a NIST invention & patent and
: > explains very well how it works. Perhaps what I and You have
: > seen is a direct consequence of applying the Smart Clock algorithm.
: >
: > Best regards
: > Ulrich
: >
: >
: > > -----Ursprungliche Nachricht-----
: > > Von: time-nuts-bounces@febo.com
: > > [mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby
: > > Gesendet: Donnerstag, 18. Juni 2009 01:51
: > > An: Discussion of precise time and frequency measurement
: > > Betreff: Re: [time-nuts] Z3805 initial behaviour after power up
: > >
: > >
: > > They do not talk about driving the frequency off, but they
: > > warn to keep
: > > the receiver locked the first 24 hours, so it can determine how the
: > > oscillator ages.  Its in the section on "holdover" (page 52
: > > of the PDF ,
: > > page 3-8 of the user guide).
: > >
: > > You may also want to search for HP An1279, it also goes over HP
: > > smartclock operation, and look for a paper titled "the Global
: > > Positioning System and HP SmartClock" by John A. Kusters.
: > >
: > > If you have a broadband connection, I can email you each
: > one, if that
: > > will help.
: > >
: > > Brian
: > >
: > > Ulrich Bangert wrote:
: > > > Brian,
: > > >
: > > > thanks for your information!
: > > >
: > > >
: > > >> algorithm.  From what I read about the smart clock, it
: > takes 5 days
: > > >> for it to complete its initial learning cycle and then its
: > > >>
: > > > continuously
: > > >
: > > >> refining.
: > > >>
: > > >
: > > > Is that from the Z3801 manual, which I have available, or
: > > do you have
: > > > any other in depth information source?
: > > >
: > > > Best regards
: > > > Ulrich
: > > >
: > > >
: > > >> -----Ursprungliche Nachricht-----
: > > >> Von: time-nuts-bounces@febo.com
: > [mailto:time-nuts-bounces@febo.com]
: > > >> Im Auftrag von Brian Kirby
: > > >> Gesendet: Mittwoch, 17. Juni 2009 16:58
: > > >> An: Discussion of precise time and frequency measurement
: > > >> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up
: > > >>
: > > >>
: > > >> Ulrich,
: > > >>
: > > >> I have two Z3801A and both of them act close to what you
: > describe
: > > >> in the first 24 hours of power up.  I believe its part of the
: > > disciplining
: > > >> algorithm.  From what I read about the smart clock, it takes
: > > >> 5 days for
: > > >> it to complete its initial learning cycle and then its
: > > continuously
: > > >> refining.
: > > >>
: > > >> Brian
: > > >>
: > > >> _______________________________________________
: > > >> time-nuts mailing list -- time-nuts@febo.com
: > > >> To unsubscribe, go to
: > > >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
: > > >> and follow the instructions there.
: > > >>
: > > >
: > > >
: > > > _______________________________________________
: > > > time-nuts mailing list -- time-nuts@febo.com
: > > > To unsubscribe, go to
: > > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
: > > > and follow the instructions there.
: > > >
: > > >
: > > _______________________________________________
: > > time-nuts mailing list -- time-nuts@febo.com
: > > To unsubscribe, go to
: > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
: > > and follow the instructions there.
: >
: >
: > _______________________________________________
: > time-nuts mailing list -- time-nuts@febo.com
: > To unsubscribe, go to
: > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
: > and follow the instructions there.
:
:
: _______________________________________________
: time-nuts mailing list -- time-nuts@febo.com
: To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
: and follow the instructions there.
:
:

In message: <PAEJKNPPIGMNDLMJLENMGEFJHNAA.frledda@verizon.net> "Francesco Ledda" <frledda@verizon.net> writes: : Aging cannot be predicted! If it could be predicetd, there would be no : aging. Actually, if aging could be predicted perfectly, there'd be perfect holdover (which I think is saying the same thing). Aging can be predicted imperfectly in the short term, but the amount of imperfect grows with the time you're without data... Warner : : -----Original Message----- : From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On : Behalf Of Ulrich Bangert : Sent: Thursday, June 18, 2009 3:04 AM : To: 'Discussion of precise time and frequency measurement' : Subject: [time-nuts] Z3805 initial behaviour after power up revised : : : Gents, : : one of the papers suggested by Brian says: : ------------------------------------------- : SmartClock monitors the frequency control variable of the internal : oscillator while it is locked to the external reference. This gives a : measure of the frequency difference between the internal oscillator, if it : is free-running, and the external reference over time. The resulting : measurements include the effects of random noise in the oscillator, the : measurement circuitry, and any noise in the external reference as well as : any aging and environmental effects in the oscillator. From this : information, SmartClock makes a continuous prediction of clock error over : time. : ------------------------------------------- : : The big question is: If the EFC signal includes this all information, how : does the algo manage to extract the individual informations? After having a : look at the PPS TI and the EFC of my Z3805 I am beginning to get a clue of : it (the quirks on the EFC signal heve been removed): : : The SmartClock algo seems to set the loop time constants to a large value in : the beginning. In other words: It measures the overall oscillator frequency : influencing effects with a lowpass filter applied that has a very low cutoff : frequency. Only effects which's frequency are sufficient lower then the : cutoff pass the filter. Seems as if the SmartClock uses this filter setting : to exclude any noise related effect and any environmental effect and make : the measurement see ONLY the oscillator's aging. : : Once it has learned enough about the aging process it will be able to : predict aging. Having reached this state it will be able to apply his : knowledge about aging to further TI measurements and subtract the "aging : part" of it, leaving mostly the environmental (which is mostly temperature) : part. : : Having a model for the aging the Algo can now model the temperature : dependence. However, a precondition for this were that the measurement sees : the temperature effects and that the ambient temperature is measured : independently. For that reason I expect the Algo will reduce its loop time : constant within the next time to make the temperature effect get through the : lowpass. Any bets on that? : : Best regards : Ulrich Bangert : : > -----Ursprungliche Nachricht----- : > Von: time-nuts-bounces@febo.com : > [mailto:time-nuts-bounces@febo.com] Im Auftrag von Ulrich Bangert : > Gesendet: Donnerstag, 18. Juni 2009 08:29 : > An: 'Discussion of precise time and frequency measurement' : > Betreff: Re: [time-nuts] Z3805 initial behaviour after power up : > : > : > Brian, : > : > > You may also want to search for HP An1279, it also goes over HP : > > smartclock operation, and look for a paper titled "the Global : > > Positioning System and HP SmartClock" by John A. Kusters. : > : > In the meantime I have not only found these but also "Smart : > Clock: A New Time" by David Allan et al which shows that : > "Smart Clock" is originally a NIST invention & patent and : > explains very well how it works. Perhaps what I and You have : > seen is a direct consequence of applying the Smart Clock algorithm. : > : > Best regards : > Ulrich : > : > : > > -----Ursprungliche Nachricht----- : > > Von: time-nuts-bounces@febo.com : > > [mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby : > > Gesendet: Donnerstag, 18. Juni 2009 01:51 : > > An: Discussion of precise time and frequency measurement : > > Betreff: Re: [time-nuts] Z3805 initial behaviour after power up : > > : > > : > > They do not talk about driving the frequency off, but they : > > warn to keep : > > the receiver locked the first 24 hours, so it can determine how the : > > oscillator ages. Its in the section on "holdover" (page 52 : > > of the PDF , : > > page 3-8 of the user guide). : > > : > > You may also want to search for HP An1279, it also goes over HP : > > smartclock operation, and look for a paper titled "the Global : > > Positioning System and HP SmartClock" by John A. Kusters. : > > : > > If you have a broadband connection, I can email you each : > one, if that : > > will help. : > > : > > Brian : > > : > > Ulrich Bangert wrote: : > > > Brian, : > > > : > > > thanks for your information! : > > > : > > > : > > >> algorithm. From what I read about the smart clock, it : > takes 5 days : > > >> for it to complete its initial learning cycle and then its : > > >> : > > > continuously : > > > : > > >> refining. : > > >> : > > > : > > > Is that from the Z3801 manual, which I have available, or : > > do you have : > > > any other in depth information source? : > > > : > > > Best regards : > > > Ulrich : > > > : > > > : > > >> -----Ursprungliche Nachricht----- : > > >> Von: time-nuts-bounces@febo.com : > [mailto:time-nuts-bounces@febo.com] : > > >> Im Auftrag von Brian Kirby : > > >> Gesendet: Mittwoch, 17. Juni 2009 16:58 : > > >> An: Discussion of precise time and frequency measurement : > > >> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up : > > >> : > > >> : > > >> Ulrich, : > > >> : > > >> I have two Z3801A and both of them act close to what you : > describe : > > >> in the first 24 hours of power up. I believe its part of the : > > disciplining : > > >> algorithm. From what I read about the smart clock, it takes : > > >> 5 days for : > > >> it to complete its initial learning cycle and then its : > > continuously : > > >> refining. : > > >> : > > >> Brian : > > >> : > > >> _______________________________________________ : > > >> time-nuts mailing list -- time-nuts@febo.com : > > >> To unsubscribe, go to : > > >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts : > > >> and follow the instructions there. : > > >> : > > > : > > > : > > > _______________________________________________ : > > > time-nuts mailing list -- time-nuts@febo.com : > > > To unsubscribe, go to : > > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts : > > > and follow the instructions there. : > > > : > > > : > > _______________________________________________ : > > time-nuts mailing list -- time-nuts@febo.com : > > To unsubscribe, go to : > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts : > > and follow the instructions there. : > : > : > _______________________________________________ : > time-nuts mailing list -- time-nuts@febo.com : > To unsubscribe, go to : > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts : > and follow the instructions there. : : : _______________________________________________ : time-nuts mailing list -- time-nuts@febo.com : To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts : and follow the instructions there. : :
BG
Bruce Griffiths
Fri, Jun 19, 2009 4:53 AM

For the benefit of HP5370A and/or HP5370B users I have added a webpage
detailing the characteristics and diagnosis of a fault that my HP5370A
recently developed.
With the 10MHz rear panel output connected to both inputs it developed
an overange error when set to measure the time delay between the START
and STOP inputs.
Details are on the  web page:

http://www.ko4bb.com/~bruce/HP5370_Faults.html
http://www.ko4bb.com/%7Ebruce/HP5370_Faults.html

At Didier's suggestion I've also included a link detailing the imaging
setup and camera settings etc used to acquire the chip image on the
above page.
Whilst it is possible to obtain good quality images with less precise
multiaxis positioner than those used, at least it is very easy using
such equipment
The chip image confirms the diagnosis inferred from measurements of
various signals within the 5370A.

Replacing A22U21 cured the problem,

Bruce

For the benefit of HP5370A and/or HP5370B users I have added a webpage detailing the characteristics and diagnosis of a fault that my HP5370A recently developed. With the 10MHz rear panel output connected to both inputs it developed an overange error when set to measure the time delay between the START and STOP inputs. Details are on the web page: http://www.ko4bb.com/~bruce/HP5370_Faults.html <http://www.ko4bb.com/%7Ebruce/HP5370_Faults.html> At Didier's suggestion I've also included a link detailing the imaging setup and camera settings etc used to acquire the chip image on the above page. Whilst it is possible to obtain good quality images with less precise multiaxis positioner than those used, at least it is very easy using such equipment The chip image confirms the diagnosis inferred from measurements of various signals within the 5370A. Replacing A22U21 cured the problem, Bruce
DF
David Forbes
Fri, Jun 19, 2009 6:12 AM

At 4:53 PM +1200 6/19/09, Bruce Griffiths wrote:

Bruce,

This sounds like a good excuse to learn how the wire bonder in the
office next to mine works.

--

--David Forbes, Tucson, AZ
http://www.cathodecorner.com/

At 4:53 PM +1200 6/19/09, Bruce Griffiths wrote: > >http://www.ko4bb.com/~bruce/HP5370_Faults.html ><http://www.ko4bb.com/%7Ebruce/HP5370_Faults.html> > > >Replacing A22U21 cured the problem, > >Bruce Bruce, This sounds like a good excuse to learn how the wire bonder in the office next to mine works. -- --David Forbes, Tucson, AZ http://www.cathodecorner.com/
R
Rex
Fri, Jun 19, 2009 7:46 AM

Bruce,

I've never seen you post a new thread as a response to an existing,
unrelated thread, as you've done here. You are not the first to do this,
but I thought you might be aware of these subtleties.

In the spirit of quasi-related posts...

On the 5370A part, I recently bought one cheap and it sort of works but
seems very noisey in its lower order digit readings; clean external ref
doesn't help. I haven't looked at obvious things like the power supply
yet, but I'm curious if the group has any thoughts of common problems to
look for.

Bruce Griffiths wrote:

For the benefit of HP5370A and/or HP5370B users I have added a webpage
detailing the characteristics and diagnosis of a fault that my HP5370A
recently developed.
With the 10MHz rear panel output connected to both inputs it developed
an overange error when set to measure the time delay between the START
and STOP inputs.
Details are on the  web page:

http://www.ko4bb.com/~bruce/HP5370_Faults.html
http://www.ko4bb.com/%7Ebruce/HP5370_Faults.html

At Didier's suggestion I've also included a link detailing the imaging
setup and camera settings etc used to acquire the chip image on the
above page.
Whilst it is possible to obtain good quality images with less precise
multiaxis positioner than those used, at least it is very easy using
such equipment
The chip image confirms the diagnosis inferred from measurements of
various signals within the 5370A.

Replacing A22U21 cured the problem,

Bruce


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Bruce, I've never seen you post a new thread as a response to an existing, unrelated thread, as you've done here. You are not the first to do this, but I thought you might be aware of these subtleties. In the spirit of quasi-related posts... On the 5370A part, I recently bought one cheap and it sort of works but seems very noisey in its lower order digit readings; clean external ref doesn't help. I haven't looked at obvious things like the power supply yet, but I'm curious if the group has any thoughts of common problems to look for. Bruce Griffiths wrote: >For the benefit of HP5370A and/or HP5370B users I have added a webpage >detailing the characteristics and diagnosis of a fault that my HP5370A >recently developed. >With the 10MHz rear panel output connected to both inputs it developed >an overange error when set to measure the time delay between the START >and STOP inputs. >Details are on the web page: > >http://www.ko4bb.com/~bruce/HP5370_Faults.html ><http://www.ko4bb.com/%7Ebruce/HP5370_Faults.html> > >At Didier's suggestion I've also included a link detailing the imaging >setup and camera settings etc used to acquire the chip image on the >above page. >Whilst it is possible to obtain good quality images with less precise >multiaxis positioner than those used, at least it is very easy using >such equipment >The chip image confirms the diagnosis inferred from measurements of >various signals within the 5370A. > >Replacing A22U21 cured the problem, > >Bruce > > >_______________________________________________ >time-nuts mailing list -- time-nuts@febo.com >To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >and follow the instructions there. > > > >
BG
Bruce Griffiths
Fri, Jun 19, 2009 8:09 AM

Bruce

Rex wrote:

Bruce,

I've never seen you post a new thread as a response to an existing,
unrelated thread, as you've done here. You are not the first to do
this, but I thought you might be aware of these subtleties.

I was being lazy.
How can you tell, looking at the message source I wasn't able to see any
indication of this?

In the spirit of quasi-related posts...

On the 5370A part, I recently bought one cheap and it sort of works
but seems very noisey in its lower order digit readings; clean
external ref doesn't help. I haven't looked at obvious things like the
power supply yet, but I'm curious if the group has any thoughts of
common problems to look for.

One cause of this can be misalignment of the 10MHz to 200MHz frequency
multiplier chain filters.
Another thing to look at (if you are using the rear panel 10MHz output)
is the signal may not be clean enough in which case checking that the
output filter is working properly and/or using an external bandpass
filter may be necessary to cleanup the signal sufficiently.
What is the indicated std deviation of the delay between the STAR and
STOP inputs when connecting the rear panel 10MHz signal to both START
and STOP inputs using the STAR COM position of the channel mode switch?

Bruce Griffiths wrote:

For the benefit of HP5370A and/or HP5370B users I have added a webpage
detailing the characteristics and diagnosis of a fault that my HP5370A
recently developed.
With the 10MHz rear panel output connected to both inputs it developed
an overange error when set to measure the time delay between the START
and STOP inputs.
Details are on the  web page:

http://www.ko4bb.com/~bruce/HP5370_Faults.html
http://www.ko4bb.com/%7Ebruce/HP5370_Faults.html

At Didier's suggestion I've also included a link detailing the imaging
setup and camera settings etc used to acquire the chip image on the
above page.
Whilst it is possible to obtain good quality images with less precise
multiaxis positioner than those used, at least it is very easy using
such equipment
The chip image confirms the diagnosis inferred from measurements of
various signals within the 5370A.

Replacing A22U21 cured the problem,

Bruce


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Bruce

Bruce Rex wrote: > Bruce, > > I've never seen you post a new thread as a response to an existing, > unrelated thread, as you've done here. You are not the first to do > this, but I thought you might be aware of these subtleties. > I was being lazy. How can you tell, looking at the message source I wasn't able to see any indication of this? > In the spirit of quasi-related posts... > > On the 5370A part, I recently bought one cheap and it sort of works > but seems very noisey in its lower order digit readings; clean > external ref doesn't help. I haven't looked at obvious things like the > power supply yet, but I'm curious if the group has any thoughts of > common problems to look for. > One cause of this can be misalignment of the 10MHz to 200MHz frequency multiplier chain filters. Another thing to look at (if you are using the rear panel 10MHz output) is the signal may not be clean enough in which case checking that the output filter is working properly and/or using an external bandpass filter may be necessary to cleanup the signal sufficiently. What is the indicated std deviation of the delay between the STAR and STOP inputs when connecting the rear panel 10MHz signal to both START and STOP inputs using the STAR COM position of the channel mode switch? > > Bruce Griffiths wrote: > >> For the benefit of HP5370A and/or HP5370B users I have added a webpage >> detailing the characteristics and diagnosis of a fault that my HP5370A >> recently developed. >> With the 10MHz rear panel output connected to both inputs it developed >> an overange error when set to measure the time delay between the START >> and STOP inputs. >> Details are on the web page: >> >> http://www.ko4bb.com/~bruce/HP5370_Faults.html >> <http://www.ko4bb.com/%7Ebruce/HP5370_Faults.html> >> >> At Didier's suggestion I've also included a link detailing the imaging >> setup and camera settings etc used to acquire the chip image on the >> above page. >> Whilst it is possible to obtain good quality images with less precise >> multiaxis positioner than those used, at least it is very easy using >> such equipment >> The chip image confirms the diagnosis inferred from measurements of >> various signals within the 5370A. >> >> Replacing A22U21 cured the problem, >> >> Bruce >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> >> >> >> > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > Bruce
JM
John Miles
Fri, Jun 19, 2009 8:10 AM

On the 5370A part, I recently bought one cheap and it sort of works but
seems very noisey in its lower order digit readings; clean external ref
doesn't help. I haven't looked at obvious things like the power supply
yet, but I'm curious if the group has any thoughts of common problems to
look for.

What mode are you observing the noise in?  Noisy LSDs on a 5370 are normal
in many cases.

-- john, KE5FX

> On the 5370A part, I recently bought one cheap and it sort of works but > seems very noisey in its lower order digit readings; clean external ref > doesn't help. I haven't looked at obvious things like the power supply > yet, but I'm curious if the group has any thoughts of common problems to > look for. What mode are you observing the noise in? Noisy LSDs on a 5370 are normal in many cases. -- john, KE5FX
R
Rex
Fri, Jun 19, 2009 9:16 AM

Bruce Griffiths wrote:

Bruce

Rex wrote:

Bruce,

I've never seen you post a new thread as a response to an existing,
unrelated thread, as you've done here. You are not the first to do
this, but I thought you might be aware of these subtleties.

I was being lazy.
How can you tell, looking at the message source I wasn't able to see any
indication of this?

I don't know enough about what is in the headers to explain the posting
issue -- one of the entries ties back to the previous message, I think.
The way I can tell is reading with the Thunderbird mail program, the new
message shows up still under the earlier thread but with the new
subject. I think I have seen similar with other mail programs.

Thanks for the reply to my 5370A question. I'll study on it. Since the
box sort of works, I think it should be fixable.

Bruce Griffiths wrote: >Bruce > >Rex wrote: > > >>Bruce, >> >>I've never seen you post a new thread as a response to an existing, >>unrelated thread, as you've done here. You are not the first to do >>this, but I thought you might be aware of these subtleties. >> >> >> > >I was being lazy. >How can you tell, looking at the message source I wasn't able to see any >indication of this? > > > I don't know enough about what is in the headers to explain the posting issue -- one of the entries ties back to the previous message, I think. The way I can tell is reading with the Thunderbird mail program, the new message shows up still under the earlier thread but with the new subject. I think I have seen similar with other mail programs. Thanks for the reply to my 5370A question. I'll study on it. Since the box sort of works, I think it should be fixable.
R
Rex
Fri, Jun 19, 2009 9:32 AM

John Miles wrote:

On the 5370A part, I recently bought one cheap and it sort of works but
seems very noisey in its lower order digit readings; clean external ref
doesn't help. I haven't looked at obvious things like the power supply
yet, but I'm curious if the group has any thoughts of common problems to
look for.

What mode are you observing the noise in?  Noisy LSDs on a 5370 are normal
in many cases.

-- john, KE5FX

I have a 5334B and looking at the same signal it seems more stable. From
memory (I haven't tried it recently) the 5370 had 3 or 4 unstable low
order digits. Even selecting high periods or sample size didn't seem to
remove noise out of two or three digits.

I never used a 5370 before, but this doesn't seem right. I get several
random digits even looking at it's own clean 10 MHz external clock. That
shouldn't be, should it?

John Miles wrote: >>On the 5370A part, I recently bought one cheap and it sort of works but >>seems very noisey in its lower order digit readings; clean external ref >>doesn't help. I haven't looked at obvious things like the power supply >>yet, but I'm curious if the group has any thoughts of common problems to >>look for. >> >> > >What mode are you observing the noise in? Noisy LSDs on a 5370 are normal >in many cases. > >-- john, KE5FX > > > I have a 5334B and looking at the same signal it seems more stable. From memory (I haven't tried it recently) the 5370 had 3 or 4 unstable low order digits. Even selecting high periods or sample size didn't seem to remove noise out of two or three digits. I never used a 5370 before, but this doesn't seem right. I get several random digits even looking at it's own clean 10 MHz external clock. That shouldn't be, should it?
R
Rex
Fri, Jun 19, 2009 9:53 AM

Bruce Griffiths wrote:

What is the indicated std deviation of the delay between the STAR and
STOP inputs when connecting the rear panel 10MHz signal to both START
and STOP inputs using the STAR COM position of the channel mode switch?

Bruce,

I don't think I understand exactly how to measure this.

First, let me say that this is a very early 5370A. It doesn't even have
a 10544, it has the early crappy reference. Hence, I have been feeding
it a better signal to the external ref input.

Now, I am assuming you are talking about measuring a version of this
reference signal. I think you are saying, put the front panel switch to
START COMMON. Are you saying to feed in phase inputs to both Start and
Stop inputs? I thought Start Common tied them together. Then I have no
clue how to gather a reading for std deviation.

Sorry if I am being dense. This counter, though very old, is new to me.

Thanks for the help and suggestions.

Bruce Griffiths wrote: >What is the indicated std deviation of the delay between the STAR and >STOP inputs when connecting the rear panel 10MHz signal to both START >and STOP inputs using the STAR COM position of the channel mode switch? > > Bruce, I don't think I understand exactly how to measure this. First, let me say that this is a very early 5370A. It doesn't even have a 10544, it has the early crappy reference. Hence, I have been feeding it a better signal to the external ref input. Now, I am assuming you are talking about measuring a version of this reference signal. I think you are saying, put the front panel switch to START COMMON. Are you saying to feed in phase inputs to both Start and Stop inputs? I thought Start Common tied them together. Then I have no clue how to gather a reading for std deviation. Sorry if I am being dense. This counter, though very old, is new to me. Thanks for the help and suggestions.
JM
John Miles
Fri, Jun 19, 2009 10:07 AM

I have a 5334B and looking at the same signal it seems more stable. From
memory (I haven't tried it recently) the 5370 had 3 or 4 unstable low
order digits. Even selecting high periods or sample size didn't seem to
remove noise out of two or three digits.

I never used a 5370 before, but this doesn't seem right. I get several
random digits even looking at it's own clean 10 MHz external clock. That
shouldn't be, should it?

The procedure Bruce was describing is basically the jitter test from the
manual, at http://cp.literature.agilent.com/litweb/pdf/05370-90031.pdf (this
is actually the same copy that David Kirkby scanned a few years ago, hosted
by Agilent.)  Follow the directions on page 3-11 and 3-12, and post the
results, especially from step 15.  This jitter figure is usually well under
50 ps in a 5370 that's working properly.

-- john, KE5FX

> I have a 5334B and looking at the same signal it seems more stable. From > memory (I haven't tried it recently) the 5370 had 3 or 4 unstable low > order digits. Even selecting high periods or sample size didn't seem to > remove noise out of two or three digits. > > I never used a 5370 before, but this doesn't seem right. I get several > random digits even looking at it's own clean 10 MHz external clock. That > shouldn't be, should it? The procedure Bruce was describing is basically the jitter test from the manual, at http://cp.literature.agilent.com/litweb/pdf/05370-90031.pdf (this is actually the same copy that David Kirkby scanned a few years ago, hosted by Agilent.) Follow the directions on page 3-11 and 3-12, and post the results, especially from step 15. This jitter figure is usually well under 50 ps in a 5370 that's working properly. -- john, KE5FX
R
Rex
Fri, Jun 19, 2009 10:40 AM

John Miles wrote:

I have a 5334B and looking at the same signal it seems more stable. From
memory (I haven't tried it recently) the 5370 had 3 or 4 unstable low
order digits. Even selecting high periods or sample size didn't seem to
remove noise out of two or three digits.

I never used a 5370 before, but this doesn't seem right. I get several
random digits even looking at it's own clean 10 MHz external clock. That
shouldn't be, should it?

The procedure Bruce was describing is basically the jitter test from the
manual, at http://cp.literature.agilent.com/litweb/pdf/05370-90031.pdf (this
is actually the same copy that David Kirkby scanned a few years ago, hosted
by Agilent.)  Follow the directions on page 3-11 and 3-12, and post the
results, especially from step 15.  This jitter figure is usually well under
50 ps in a 5370 that's working properly.

-- john, KE5FX

Thanks. I think I did the beginning of that, but will need to do it
again to get anything accurate to post. Will do.

John Miles wrote: >>I have a 5334B and looking at the same signal it seems more stable. From >>memory (I haven't tried it recently) the 5370 had 3 or 4 unstable low >>order digits. Even selecting high periods or sample size didn't seem to >>remove noise out of two or three digits. >> >>I never used a 5370 before, but this doesn't seem right. I get several >>random digits even looking at it's own clean 10 MHz external clock. That >>shouldn't be, should it? >> >> > >The procedure Bruce was describing is basically the jitter test from the >manual, at http://cp.literature.agilent.com/litweb/pdf/05370-90031.pdf (this >is actually the same copy that David Kirkby scanned a few years ago, hosted >by Agilent.) Follow the directions on page 3-11 and 3-12, and post the >results, especially from step 15. This jitter figure is usually well under >50 ps in a 5370 that's working properly. > >-- john, KE5FX > > > > > Thanks. I think I did the beginning of that, but will need to do it again to get anything accurate to post. Will do.
BG
Bruce Griffiths
Fri, Jun 19, 2009 10:46 AM

Rex wrote:

Bruce Griffiths wrote:

What is the indicated std deviation of the delay between the STAR and
STOP inputs when connecting the rear panel 10MHz signal to both START
and STOP inputs using the STAR COM position of the channel mode switch?

Bruce,

I don't think I understand exactly how to measure this.

First, let me say that this is a very early 5370A. It doesn't even
have a 10544, it has the early crappy reference. Hence, I have been
feeding it a better signal to the external ref input.

Now, I am assuming you are talking about measuring a version of this
reference signal. I think you are saying, put the front panel switch
to START COMMON. Are you saying to feed in phase inputs to both Start
and Stop inputs? I thought Start Common tied them together. Then I
have no clue how to gather a reading for std deviation.

Yes, I should have given clearer instructions, just connect the 10MHz
out to  the START input with the swich selecting both inputs as COMMON.
Even with the ECL gate based crystal oscillator (my 5370A has one as
well even though its not a very early one, HP seemed to jump back and
forth between having a cheap oscillator and having a 10544A as standard)
the standard deviation should be less than 100ps. Typically something
like 15ps. With an external 10MHz signal connected to both inputs in
COMMON the STD deviation will be a little larger perhaps 35ps or so.

Sorry if I am being dense. This counter, though very old, is new to me.

Thanks for the help and suggestions.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Bruce

Rex wrote: > Bruce Griffiths wrote: > >> What is the indicated std deviation of the delay between the STAR and >> STOP inputs when connecting the rear panel 10MHz signal to both START >> and STOP inputs using the STAR COM position of the channel mode switch? >> >> > Bruce, > > I don't think I understand exactly how to measure this. > > First, let me say that this is a very early 5370A. It doesn't even > have a 10544, it has the early crappy reference. Hence, I have been > feeding it a better signal to the external ref input. > > Now, I am assuming you are talking about measuring a version of this > reference signal. I think you are saying, put the front panel switch > to START COMMON. Are you saying to feed in phase inputs to both Start > and Stop inputs? I thought Start Common tied them together. Then I > have no clue how to gather a reading for std deviation. > Yes, I should have given clearer instructions, just connect the 10MHz out to the START input with the swich selecting both inputs as COMMON. Even with the ECL gate based crystal oscillator (my 5370A has one as well even though its not a very early one, HP seemed to jump back and forth between having a cheap oscillator and having a 10544A as standard) the standard deviation should be less than 100ps. Typically something like 15ps. With an external 10MHz signal connected to both inputs in COMMON the STD deviation will be a little larger perhaps 35ps or so. > Sorry if I am being dense. This counter, though very old, is new to me. > > Thanks for the help and suggestions. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > Bruce
MD
Magnus Danielson
Fri, Jun 19, 2009 12:58 PM

Francesco Ledda wrote:

Aging cannot be predicted!  If it could be predicetd, there would be no
aging.

Not entierly true. Even with a perfectly known aging, only a few
oscillators would bother to estimate and correct it. This can never be
perfectly done anyway, and there is only so many things you can bring
into the model while keeping it economical. There is fairly good clues
around. If all was included and handled, it would reduce the effects.

Cheers,
Magnus

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On
Behalf Of Ulrich Bangert
Sent: Thursday, June 18, 2009 3:04 AM
To: 'Discussion of precise time and frequency measurement'
Subject: [time-nuts] Z3805 initial behaviour after power up revised

Gents,

one of the papers suggested by Brian says:

SmartClock monitors the frequency control variable of the internal
oscillator while it is locked to the external reference. This gives a
measure of the frequency difference between the internal oscillator, if it
is free-running, and the external reference over time. The resulting
measurements include the effects of random noise in the oscillator, the
measurement circuitry, and any noise in the external reference as well as
any aging and environmental effects in the oscillator. From this
information, SmartClock makes a continuous prediction of clock error over
time.

The big question is: If the EFC signal includes this all information, how
does the algo manage to extract the individual informations? After having a
look at the PPS TI and the EFC of my Z3805 I am beginning to get a clue of
it (the quirks on the EFC signal heve been removed):

The SmartClock algo seems to set the loop time constants to a large value in
the beginning. In other words: It measures the overall oscillator frequency
influencing effects with a lowpass filter applied that has a very low cutoff
frequency. Only effects which's frequency are sufficient lower then the
cutoff pass the filter. Seems as if the SmartClock uses this filter setting
to exclude any noise related effect and any environmental effect and make
the measurement see ONLY the oscillator's aging.

Once it has learned enough about the aging process it will be able to
predict aging. Having reached this state it will be able to apply his
knowledge about aging to further TI measurements and subtract the "aging
part" of it, leaving mostly the environmental (which is mostly temperature)
part.

Having a model for the aging the Algo can now model the temperature
dependence. However, a precondition for this were that the measurement sees
the temperature effects and that the ambient temperature is measured
independently. For that reason I expect the Algo will reduce its loop time
constant within the next time to make the temperature effect get through the
lowpass. Any bets on that?

Best regards
Ulrich Bangert

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Ulrich Bangert
Gesendet: Donnerstag, 18. Juni 2009 08:29
An: 'Discussion of precise time and frequency measurement'
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Brian,

You may also want to search for HP An1279, it also goes over HP
smartclock operation, and look for a paper titled "the Global
Positioning System and HP SmartClock" by John A. Kusters.

In the meantime I have not only found these but also "Smart
Clock: A New Time" by David Allan et al which shows that
"Smart Clock" is originally a NIST invention & patent and
explains very well how it works. Perhaps what I and You have
seen is a direct consequence of applying the Smart Clock algorithm.

Best regards
Ulrich

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby
Gesendet: Donnerstag, 18. Juni 2009 01:51
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

They do not talk about driving the frequency off, but they
warn to keep
the receiver locked the first 24 hours, so it can determine how the
oscillator ages.  Its in the section on "holdover" (page 52
of the PDF ,
page 3-8 of the user guide).

You may also want to search for HP An1279, it also goes over HP
smartclock operation, and look for a paper titled "the Global
Positioning System and HP SmartClock" by John A. Kusters.

If you have a broadband connection, I can email you each

one, if that

will help.

Brian

Ulrich Bangert wrote:

Brian,

thanks for your information!

algorithm.  From what I read about the smart clock, it

takes 5 days

for it to complete its initial learning cycle and then its

continuously

refining.

Is that from the Z3801 manual, which I have available, or

do you have

any other in depth information source?

Best regards
Ulrich

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com

Im Auftrag von Brian Kirby
Gesendet: Mittwoch, 17. Juni 2009 16:58
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Ulrich,

I have two Z3801A and both of them act close to what you

describe

in the first 24 hours of power up.  I believe its part of the

disciplining

algorithm.  From what I read about the smart clock, it takes
5 days for
it to complete its initial learning cycle and then its

continuously

refining.

Brian


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Francesco Ledda wrote: > Aging cannot be predicted! If it could be predicetd, there would be no > aging. > Not entierly true. Even with a perfectly known aging, only a few oscillators would bother to estimate and correct it. This can never be perfectly done anyway, and there is only so many things you can bring into the model while keeping it economical. There is fairly good clues around. If all was included and handled, it would reduce the effects. Cheers, Magnus > -----Original Message----- > From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On > Behalf Of Ulrich Bangert > Sent: Thursday, June 18, 2009 3:04 AM > To: 'Discussion of precise time and frequency measurement' > Subject: [time-nuts] Z3805 initial behaviour after power up revised > > > Gents, > > one of the papers suggested by Brian says: > ------------------------------------------- > SmartClock monitors the frequency control variable of the internal > oscillator while it is locked to the external reference. This gives a > measure of the frequency difference between the internal oscillator, if it > is free-running, and the external reference over time. The resulting > measurements include the effects of random noise in the oscillator, the > measurement circuitry, and any noise in the external reference as well as > any aging and environmental effects in the oscillator. From this > information, SmartClock makes a continuous prediction of clock error over > time. > ------------------------------------------- > > The big question is: If the EFC signal includes this all information, how > does the algo manage to extract the individual informations? After having a > look at the PPS TI and the EFC of my Z3805 I am beginning to get a clue of > it (the quirks on the EFC signal heve been removed): > > The SmartClock algo seems to set the loop time constants to a large value in > the beginning. In other words: It measures the overall oscillator frequency > influencing effects with a lowpass filter applied that has a very low cutoff > frequency. Only effects which's frequency are sufficient lower then the > cutoff pass the filter. Seems as if the SmartClock uses this filter setting > to exclude any noise related effect and any environmental effect and make > the measurement see ONLY the oscillator's aging. > > Once it has learned enough about the aging process it will be able to > predict aging. Having reached this state it will be able to apply his > knowledge about aging to further TI measurements and subtract the "aging > part" of it, leaving mostly the environmental (which is mostly temperature) > part. > > Having a model for the aging the Algo can now model the temperature > dependence. However, a precondition for this were that the measurement sees > the temperature effects and that the ambient temperature is measured > independently. For that reason I expect the Algo will reduce its loop time > constant within the next time to make the temperature effect get through the > lowpass. Any bets on that? > > Best regards > Ulrich Bangert > > >> -----Ursprungliche Nachricht----- >> Von: time-nuts-bounces@febo.com >> [mailto:time-nuts-bounces@febo.com] Im Auftrag von Ulrich Bangert >> Gesendet: Donnerstag, 18. Juni 2009 08:29 >> An: 'Discussion of precise time and frequency measurement' >> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up >> >> >> Brian, >> >> >>> You may also want to search for HP An1279, it also goes over HP >>> smartclock operation, and look for a paper titled "the Global >>> Positioning System and HP SmartClock" by John A. Kusters. >>> >> In the meantime I have not only found these but also "Smart >> Clock: A New Time" by David Allan et al which shows that >> "Smart Clock" is originally a NIST invention & patent and >> explains very well how it works. Perhaps what I and You have >> seen is a direct consequence of applying the Smart Clock algorithm. >> >> Best regards >> Ulrich >> >> >> >>> -----Ursprungliche Nachricht----- >>> Von: time-nuts-bounces@febo.com >>> [mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby >>> Gesendet: Donnerstag, 18. Juni 2009 01:51 >>> An: Discussion of precise time and frequency measurement >>> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up >>> >>> >>> They do not talk about driving the frequency off, but they >>> warn to keep >>> the receiver locked the first 24 hours, so it can determine how the >>> oscillator ages. Its in the section on "holdover" (page 52 >>> of the PDF , >>> page 3-8 of the user guide). >>> >>> You may also want to search for HP An1279, it also goes over HP >>> smartclock operation, and look for a paper titled "the Global >>> Positioning System and HP SmartClock" by John A. Kusters. >>> >>> If you have a broadband connection, I can email you each >>> >> one, if that >> >>> will help. >>> >>> Brian >>> >>> Ulrich Bangert wrote: >>> >>>> Brian, >>>> >>>> thanks for your information! >>>> >>>> >>>> >>>>> algorithm. From what I read about the smart clock, it >>>>> >> takes 5 days >> >>>>> for it to complete its initial learning cycle and then its >>>>> >>>>> >>>> continuously >>>> >>>> >>>>> refining. >>>>> >>>>> >>>> Is that from the Z3801 manual, which I have available, or >>>> >>> do you have >>> >>>> any other in depth information source? >>>> >>>> Best regards >>>> Ulrich >>>> >>>> >>>> >>>>> -----Ursprungliche Nachricht----- >>>>> Von: time-nuts-bounces@febo.com >>>>> >> [mailto:time-nuts-bounces@febo.com] >> >>>>> Im Auftrag von Brian Kirby >>>>> Gesendet: Mittwoch, 17. Juni 2009 16:58 >>>>> An: Discussion of precise time and frequency measurement >>>>> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up >>>>> >>>>> >>>>> Ulrich, >>>>> >>>>> I have two Z3801A and both of them act close to what you >>>>> >> describe >> >>>>> in the first 24 hours of power up. I believe its part of the >>>>> >>> disciplining >>> >>>>> algorithm. From what I read about the smart clock, it takes >>>>> 5 days for >>>>> it to complete its initial learning cycle and then its >>>>> >>> continuously >>> >>>>> refining. >>>>> >>>>> Brian >>>>> >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@febo.com >>>>> To unsubscribe, go to >>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>>> and follow the instructions there. >>>>> >>>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com >>>> To unsubscribe, go to >>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>> and follow the instructions there. >>>> >>>> >>>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
CH
Chuck Harris
Fri, Jun 19, 2009 1:15 PM

Bruce Griffiths wrote:

Bruce

Rex wrote:

Bruce,

I've never seen you post a new thread as a response to an existing,
unrelated thread, as you've done here. You are not the first to do
this, but I thought you might be aware of these subtleties.

I was being lazy.
How can you tell, looking at the message source I wasn't able to see any
indication of this?

The references header points back to the post you replied to.  In this
case:

4ECC7583323F41D38EFD9C7AD619FE53@athlon

-Chuck Harris

Bruce Griffiths wrote: > Bruce > > Rex wrote: >> Bruce, >> >> I've never seen you post a new thread as a response to an existing, >> unrelated thread, as you've done here. You are not the first to do >> this, but I thought you might be aware of these subtleties. >> > > I was being lazy. > How can you tell, looking at the message source I wasn't able to see any > indication of this? The references header points back to the post you replied to. In this case: <4ECC7583323F41D38EFD9C7AD619FE53@athlon> -Chuck Harris
FL
Francesco Ledda
Fri, Jun 19, 2009 2:38 PM

Variation due to environment are deterministic and are nor part of aging.
Ageing needs to measured in conditions that remove ambient induced
perturbations. I used to put crystal oscillators in high quality
environmental chambers that kept temp, humidity and pressure constant and
measure the ageing.  There are tricks that can be done to increase the aging
rate, so that we did not have to wait 20 years ;)

We know that ageing happens, but the direction of ageing cannot be
determined.
A coin toss can yield face or tail.  If the coin is perfect, the
distribution will be uniform.  If a face is 1 and a tail is -1, we can add
successive tosses and track the total number.  Even is the distribution of
tosses is uniform, the sum will walk away from 0 (random walk),cameback to 0
and then move away from 0 again.  This is a good way to model and explain
ageing. Nature is perfect, but things do not follow our math perfectly.  A
better simulation would include a "leaky integral".

The statistical analysis of ageing is tricky, since ageing doeas have a
statistical mean, and therefore standard deviation cannot be used.

This discussion brings lots of good memories back.  The good old days when
SONET synchronization was a new thing, and we were creating new technology.

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On
Behalf Of Magnus Danielson
Sent: Friday, June 19, 2009 7:59 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Z3805 initial behaviour after power up revised

Francesco Ledda wrote:

Aging cannot be predicted!  If it could be predicetd, there would be no
aging.

Not entierly true. Even with a perfectly known aging, only a few
oscillators would bother to estimate and correct it. This can never be
perfectly done anyway, and there is only so many things you can bring
into the model while keeping it economical. There is fairly good clues
around. If all was included and handled, it would reduce the effects.

Cheers,
Magnus

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On
Behalf Of Ulrich Bangert
Sent: Thursday, June 18, 2009 3:04 AM
To: 'Discussion of precise time and frequency measurement'
Subject: [time-nuts] Z3805 initial behaviour after power up revised

Gents,

one of the papers suggested by Brian says:

SmartClock monitors the frequency control variable of the internal
oscillator while it is locked to the external reference. This gives a
measure of the frequency difference between the internal oscillator, if it
is free-running, and the external reference over time. The resulting
measurements include the effects of random noise in the oscillator, the
measurement circuitry, and any noise in the external reference as well as
any aging and environmental effects in the oscillator. From this
information, SmartClock makes a continuous prediction of clock error over
time.

The big question is: If the EFC signal includes this all information, how
does the algo manage to extract the individual informations? After having

a

look at the PPS TI and the EFC of my Z3805 I am beginning to get a clue of
it (the quirks on the EFC signal heve been removed):

The SmartClock algo seems to set the loop time constants to a large value

in

the beginning. In other words: It measures the overall oscillator

frequency

influencing effects with a lowpass filter applied that has a very low

cutoff

frequency. Only effects which's frequency are sufficient lower then the
cutoff pass the filter. Seems as if the SmartClock uses this filter

setting

to exclude any noise related effect and any environmental effect and make
the measurement see ONLY the oscillator's aging.

Once it has learned enough about the aging process it will be able to
predict aging. Having reached this state it will be able to apply his
knowledge about aging to further TI measurements and subtract the "aging
part" of it, leaving mostly the environmental (which is mostly

temperature)

part.

Having a model for the aging the Algo can now model the temperature
dependence. However, a precondition for this were that the measurement

sees

the temperature effects and that the ambient temperature is measured
independently. For that reason I expect the Algo will reduce its loop time
constant within the next time to make the temperature effect get through

the

lowpass. Any bets on that?

Best regards
Ulrich Bangert

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Ulrich Bangert
Gesendet: Donnerstag, 18. Juni 2009 08:29
An: 'Discussion of precise time and frequency measurement'
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Brian,

You may also want to search for HP An1279, it also goes over HP
smartclock operation, and look for a paper titled "the Global
Positioning System and HP SmartClock" by John A. Kusters.

In the meantime I have not only found these but also "Smart
Clock: A New Time" by David Allan et al which shows that
"Smart Clock" is originally a NIST invention & patent and
explains very well how it works. Perhaps what I and You have
seen is a direct consequence of applying the Smart Clock algorithm.

Best regards
Ulrich

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby
Gesendet: Donnerstag, 18. Juni 2009 01:51
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

They do not talk about driving the frequency off, but they
warn to keep
the receiver locked the first 24 hours, so it can determine how the
oscillator ages.  Its in the section on "holdover" (page 52
of the PDF ,
page 3-8 of the user guide).

You may also want to search for HP An1279, it also goes over HP
smartclock operation, and look for a paper titled "the Global
Positioning System and HP SmartClock" by John A. Kusters.

If you have a broadband connection, I can email you each

one, if that

will help.

Brian

Ulrich Bangert wrote:

Brian,

thanks for your information!

algorithm.  From what I read about the smart clock, it

takes 5 days

for it to complete its initial learning cycle and then its

continuously

refining.

Is that from the Z3801 manual, which I have available, or

do you have

any other in depth information source?

Best regards
Ulrich

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com

Im Auftrag von Brian Kirby
Gesendet: Mittwoch, 17. Juni 2009 16:58
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Ulrich,

I have two Z3801A and both of them act close to what you

describe

in the first 24 hours of power up.  I believe its part of the

disciplining

algorithm.  From what I read about the smart clock, it takes
5 days for
it to complete its initial learning cycle and then its

continuously

refining.

Brian


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Variation due to environment are deterministic and are nor part of aging. Ageing needs to measured in conditions that remove ambient induced perturbations. I used to put crystal oscillators in high quality environmental chambers that kept temp, humidity and pressure constant and measure the ageing. There are tricks that can be done to increase the aging rate, so that we did not have to wait 20 years ;) We know that ageing happens, but the direction of ageing cannot be determined. A coin toss can yield face or tail. If the coin is perfect, the distribution will be uniform. If a face is 1 and a tail is -1, we can add successive tosses and track the total number. Even is the distribution of tosses is uniform, the sum will walk away from 0 (random walk),cameback to 0 and then move away from 0 again. This is a good way to model and explain ageing. Nature is perfect, but things do not follow our math perfectly. A better simulation would include a "leaky integral". The statistical analysis of ageing is tricky, since ageing doeas have a statistical mean, and therefore standard deviation cannot be used. This discussion brings lots of good memories back. The good old days when SONET synchronization was a new thing, and we were creating new technology. -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On Behalf Of Magnus Danielson Sent: Friday, June 19, 2009 7:59 AM To: Discussion of precise time and frequency measurement Subject: Re: [time-nuts] Z3805 initial behaviour after power up revised Francesco Ledda wrote: > Aging cannot be predicted! If it could be predicetd, there would be no > aging. > Not entierly true. Even with a perfectly known aging, only a few oscillators would bother to estimate and correct it. This can never be perfectly done anyway, and there is only so many things you can bring into the model while keeping it economical. There is fairly good clues around. If all was included and handled, it would reduce the effects. Cheers, Magnus > -----Original Message----- > From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On > Behalf Of Ulrich Bangert > Sent: Thursday, June 18, 2009 3:04 AM > To: 'Discussion of precise time and frequency measurement' > Subject: [time-nuts] Z3805 initial behaviour after power up revised > > > Gents, > > one of the papers suggested by Brian says: > ------------------------------------------- > SmartClock monitors the frequency control variable of the internal > oscillator while it is locked to the external reference. This gives a > measure of the frequency difference between the internal oscillator, if it > is free-running, and the external reference over time. The resulting > measurements include the effects of random noise in the oscillator, the > measurement circuitry, and any noise in the external reference as well as > any aging and environmental effects in the oscillator. From this > information, SmartClock makes a continuous prediction of clock error over > time. > ------------------------------------------- > > The big question is: If the EFC signal includes this all information, how > does the algo manage to extract the individual informations? After having a > look at the PPS TI and the EFC of my Z3805 I am beginning to get a clue of > it (the quirks on the EFC signal heve been removed): > > The SmartClock algo seems to set the loop time constants to a large value in > the beginning. In other words: It measures the overall oscillator frequency > influencing effects with a lowpass filter applied that has a very low cutoff > frequency. Only effects which's frequency are sufficient lower then the > cutoff pass the filter. Seems as if the SmartClock uses this filter setting > to exclude any noise related effect and any environmental effect and make > the measurement see ONLY the oscillator's aging. > > Once it has learned enough about the aging process it will be able to > predict aging. Having reached this state it will be able to apply his > knowledge about aging to further TI measurements and subtract the "aging > part" of it, leaving mostly the environmental (which is mostly temperature) > part. > > Having a model for the aging the Algo can now model the temperature > dependence. However, a precondition for this were that the measurement sees > the temperature effects and that the ambient temperature is measured > independently. For that reason I expect the Algo will reduce its loop time > constant within the next time to make the temperature effect get through the > lowpass. Any bets on that? > > Best regards > Ulrich Bangert > > >> -----Ursprungliche Nachricht----- >> Von: time-nuts-bounces@febo.com >> [mailto:time-nuts-bounces@febo.com] Im Auftrag von Ulrich Bangert >> Gesendet: Donnerstag, 18. Juni 2009 08:29 >> An: 'Discussion of precise time and frequency measurement' >> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up >> >> >> Brian, >> >> >>> You may also want to search for HP An1279, it also goes over HP >>> smartclock operation, and look for a paper titled "the Global >>> Positioning System and HP SmartClock" by John A. Kusters. >>> >> In the meantime I have not only found these but also "Smart >> Clock: A New Time" by David Allan et al which shows that >> "Smart Clock" is originally a NIST invention & patent and >> explains very well how it works. Perhaps what I and You have >> seen is a direct consequence of applying the Smart Clock algorithm. >> >> Best regards >> Ulrich >> >> >> >>> -----Ursprungliche Nachricht----- >>> Von: time-nuts-bounces@febo.com >>> [mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby >>> Gesendet: Donnerstag, 18. Juni 2009 01:51 >>> An: Discussion of precise time and frequency measurement >>> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up >>> >>> >>> They do not talk about driving the frequency off, but they >>> warn to keep >>> the receiver locked the first 24 hours, so it can determine how the >>> oscillator ages. Its in the section on "holdover" (page 52 >>> of the PDF , >>> page 3-8 of the user guide). >>> >>> You may also want to search for HP An1279, it also goes over HP >>> smartclock operation, and look for a paper titled "the Global >>> Positioning System and HP SmartClock" by John A. Kusters. >>> >>> If you have a broadband connection, I can email you each >>> >> one, if that >> >>> will help. >>> >>> Brian >>> >>> Ulrich Bangert wrote: >>> >>>> Brian, >>>> >>>> thanks for your information! >>>> >>>> >>>> >>>>> algorithm. From what I read about the smart clock, it >>>>> >> takes 5 days >> >>>>> for it to complete its initial learning cycle and then its >>>>> >>>>> >>>> continuously >>>> >>>> >>>>> refining. >>>>> >>>>> >>>> Is that from the Z3801 manual, which I have available, or >>>> >>> do you have >>> >>>> any other in depth information source? >>>> >>>> Best regards >>>> Ulrich >>>> >>>> >>>> >>>>> -----Ursprungliche Nachricht----- >>>>> Von: time-nuts-bounces@febo.com >>>>> >> [mailto:time-nuts-bounces@febo.com] >> >>>>> Im Auftrag von Brian Kirby >>>>> Gesendet: Mittwoch, 17. Juni 2009 16:58 >>>>> An: Discussion of precise time and frequency measurement >>>>> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up >>>>> >>>>> >>>>> Ulrich, >>>>> >>>>> I have two Z3801A and both of them act close to what you >>>>> >> describe >> >>>>> in the first 24 hours of power up. I believe its part of the >>>>> >>> disciplining >>> >>>>> algorithm. From what I read about the smart clock, it takes >>>>> 5 days for >>>>> it to complete its initial learning cycle and then its >>>>> >>> continuously >>> >>>>> refining. >>>>> >>>>> Brian >>>>> >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@febo.com >>>>> To unsubscribe, go to >>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>>> and follow the instructions there. >>>>> >>>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com >>>> To unsubscribe, go to >>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>> and follow the instructions there. >>>> >>>> >>>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
FL
Francesco Ledda
Fri, Jun 19, 2009 3:57 PM

Correction...

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On
Behalf Of Francesco Ledda
Sent: Friday, June 19, 2009 9:38 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Z3805 initial behavior after power up revised

Variation due to environment are deterministic and are nor part of aging.
Ageing needs to measured in conditions that remove ambient induced
perturbations. I used to put crystal oscillators in high quality
environmental chambers that kept temp, humidity and pressure constant and
measure the ageing.  There are tricks that can be done to increase the aging
rate, so that we did not have to wait 20 years ;)

We know that ageing happens, but the direction of ageing cannot be
determined.
A coin toss can yield face or tail.  If the coin is perfect, the
distribution will be uniform.  If a face is 1 and a tail is -1, we can add
successive tosses and track the total number.  Even is the distribution of
tosses is uniform, the sum will walk away from 0 (random walk),cameback to 0
and then move away from 0 again.  This is a good way to model and explain
ageing. Nature is perfect, but things do not follow our math perfectly.  A
better simulation would include a "leaky integral".

The statistical analysis of ageing is tricky, since ageing doeas have NOT a
statistical mean, and therefore standard deviation cannot be used.

This discussion brings lots of good memories back.  The good old days when
SONET synchronization was a new thing, and we were creating new technology.

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On
Behalf Of Magnus Danielson
Sent: Friday, June 19, 2009 7:59 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Z3805 initial behaviour after power up revised

Francesco Ledda wrote:

Aging cannot be predicted!  If it could be predicetd, there would be no
aging.

Not entierly true. Even with a perfectly known aging, only a few
oscillators would bother to estimate and correct it. This can never be
perfectly done anyway, and there is only so many things you can bring
into the model while keeping it economical. There is fairly good clues
around. If all was included and handled, it would reduce the effects.

Cheers,
Magnus

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On
Behalf Of Ulrich Bangert
Sent: Thursday, June 18, 2009 3:04 AM
To: 'Discussion of precise time and frequency measurement'
Subject: [time-nuts] Z3805 initial behaviour after power up revised

Gents,

one of the papers suggested by Brian says:

SmartClock monitors the frequency control variable of the internal
oscillator while it is locked to the external reference. This gives a
measure of the frequency difference between the internal oscillator, if it
is free-running, and the external reference over time. The resulting
measurements include the effects of random noise in the oscillator, the
measurement circuitry, and any noise in the external reference as well as
any aging and environmental effects in the oscillator. From this
information, SmartClock makes a continuous prediction of clock error over
time.

The big question is: If the EFC signal includes this all information, how
does the algo manage to extract the individual informations? After having

a

look at the PPS TI and the EFC of my Z3805 I am beginning to get a clue of
it (the quirks on the EFC signal heve been removed):

The SmartClock algo seems to set the loop time constants to a large value

in

the beginning. In other words: It measures the overall oscillator

frequency

influencing effects with a lowpass filter applied that has a very low

cutoff

frequency. Only effects which's frequency are sufficient lower then the
cutoff pass the filter. Seems as if the SmartClock uses this filter

setting

to exclude any noise related effect and any environmental effect and make
the measurement see ONLY the oscillator's aging.

Once it has learned enough about the aging process it will be able to
predict aging. Having reached this state it will be able to apply his
knowledge about aging to further TI measurements and subtract the "aging
part" of it, leaving mostly the environmental (which is mostly

temperature)

part.

Having a model for the aging the Algo can now model the temperature
dependence. However, a precondition for this were that the measurement

sees

the temperature effects and that the ambient temperature is measured
independently. For that reason I expect the Algo will reduce its loop time
constant within the next time to make the temperature effect get through

the

lowpass. Any bets on that?

Best regards
Ulrich Bangert

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Ulrich Bangert
Gesendet: Donnerstag, 18. Juni 2009 08:29
An: 'Discussion of precise time and frequency measurement'
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Brian,

You may also want to search for HP An1279, it also goes over HP
smartclock operation, and look for a paper titled "the Global
Positioning System and HP SmartClock" by John A. Kusters.

In the meantime I have not only found these but also "Smart
Clock: A New Time" by David Allan et al which shows that
"Smart Clock" is originally a NIST invention & patent and
explains very well how it works. Perhaps what I and You have
seen is a direct consequence of applying the Smart Clock algorithm.

Best regards
Ulrich

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby
Gesendet: Donnerstag, 18. Juni 2009 01:51
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

They do not talk about driving the frequency off, but they
warn to keep
the receiver locked the first 24 hours, so it can determine how the
oscillator ages.  Its in the section on "holdover" (page 52
of the PDF ,
page 3-8 of the user guide).

You may also want to search for HP An1279, it also goes over HP
smartclock operation, and look for a paper titled "the Global
Positioning System and HP SmartClock" by John A. Kusters.

If you have a broadband connection, I can email you each

one, if that

will help.

Brian

Ulrich Bangert wrote:

Brian,

thanks for your information!

algorithm.  From what I read about the smart clock, it

takes 5 days

for it to complete its initial learning cycle and then its

continuously

refining.

Is that from the Z3801 manual, which I have available, or

do you have

any other in depth information source?

Best regards
Ulrich

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com

Im Auftrag von Brian Kirby
Gesendet: Mittwoch, 17. Juni 2009 16:58
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] Z3805 initial behaviour after power up

Ulrich,

I have two Z3801A and both of them act close to what you

describe

in the first 24 hours of power up.  I believe its part of the

disciplining

algorithm.  From what I read about the smart clock, it takes
5 days for
it to complete its initial learning cycle and then its

continuously

refining.

Brian


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Correction... -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On Behalf Of Francesco Ledda Sent: Friday, June 19, 2009 9:38 AM To: Discussion of precise time and frequency measurement Subject: Re: [time-nuts] Z3805 initial behavior after power up revised Variation due to environment are deterministic and are nor part of aging. Ageing needs to measured in conditions that remove ambient induced perturbations. I used to put crystal oscillators in high quality environmental chambers that kept temp, humidity and pressure constant and measure the ageing. There are tricks that can be done to increase the aging rate, so that we did not have to wait 20 years ;) We know that ageing happens, but the direction of ageing cannot be determined. A coin toss can yield face or tail. If the coin is perfect, the distribution will be uniform. If a face is 1 and a tail is -1, we can add successive tosses and track the total number. Even is the distribution of tosses is uniform, the sum will walk away from 0 (random walk),cameback to 0 and then move away from 0 again. This is a good way to model and explain ageing. Nature is perfect, but things do not follow our math perfectly. A better simulation would include a "leaky integral". The statistical analysis of ageing is tricky, since ageing doeas have NOT a statistical mean, and therefore standard deviation cannot be used. This discussion brings lots of good memories back. The good old days when SONET synchronization was a new thing, and we were creating new technology. -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On Behalf Of Magnus Danielson Sent: Friday, June 19, 2009 7:59 AM To: Discussion of precise time and frequency measurement Subject: Re: [time-nuts] Z3805 initial behaviour after power up revised Francesco Ledda wrote: > Aging cannot be predicted! If it could be predicetd, there would be no > aging. > Not entierly true. Even with a perfectly known aging, only a few oscillators would bother to estimate and correct it. This can never be perfectly done anyway, and there is only so many things you can bring into the model while keeping it economical. There is fairly good clues around. If all was included and handled, it would reduce the effects. Cheers, Magnus > -----Original Message----- > From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On > Behalf Of Ulrich Bangert > Sent: Thursday, June 18, 2009 3:04 AM > To: 'Discussion of precise time and frequency measurement' > Subject: [time-nuts] Z3805 initial behaviour after power up revised > > > Gents, > > one of the papers suggested by Brian says: > ------------------------------------------- > SmartClock monitors the frequency control variable of the internal > oscillator while it is locked to the external reference. This gives a > measure of the frequency difference between the internal oscillator, if it > is free-running, and the external reference over time. The resulting > measurements include the effects of random noise in the oscillator, the > measurement circuitry, and any noise in the external reference as well as > any aging and environmental effects in the oscillator. From this > information, SmartClock makes a continuous prediction of clock error over > time. > ------------------------------------------- > > The big question is: If the EFC signal includes this all information, how > does the algo manage to extract the individual informations? After having a > look at the PPS TI and the EFC of my Z3805 I am beginning to get a clue of > it (the quirks on the EFC signal heve been removed): > > The SmartClock algo seems to set the loop time constants to a large value in > the beginning. In other words: It measures the overall oscillator frequency > influencing effects with a lowpass filter applied that has a very low cutoff > frequency. Only effects which's frequency are sufficient lower then the > cutoff pass the filter. Seems as if the SmartClock uses this filter setting > to exclude any noise related effect and any environmental effect and make > the measurement see ONLY the oscillator's aging. > > Once it has learned enough about the aging process it will be able to > predict aging. Having reached this state it will be able to apply his > knowledge about aging to further TI measurements and subtract the "aging > part" of it, leaving mostly the environmental (which is mostly temperature) > part. > > Having a model for the aging the Algo can now model the temperature > dependence. However, a precondition for this were that the measurement sees > the temperature effects and that the ambient temperature is measured > independently. For that reason I expect the Algo will reduce its loop time > constant within the next time to make the temperature effect get through the > lowpass. Any bets on that? > > Best regards > Ulrich Bangert > > >> -----Ursprungliche Nachricht----- >> Von: time-nuts-bounces@febo.com >> [mailto:time-nuts-bounces@febo.com] Im Auftrag von Ulrich Bangert >> Gesendet: Donnerstag, 18. Juni 2009 08:29 >> An: 'Discussion of precise time and frequency measurement' >> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up >> >> >> Brian, >> >> >>> You may also want to search for HP An1279, it also goes over HP >>> smartclock operation, and look for a paper titled "the Global >>> Positioning System and HP SmartClock" by John A. Kusters. >>> >> In the meantime I have not only found these but also "Smart >> Clock: A New Time" by David Allan et al which shows that >> "Smart Clock" is originally a NIST invention & patent and >> explains very well how it works. Perhaps what I and You have >> seen is a direct consequence of applying the Smart Clock algorithm. >> >> Best regards >> Ulrich >> >> >> >>> -----Ursprungliche Nachricht----- >>> Von: time-nuts-bounces@febo.com >>> [mailto:time-nuts-bounces@febo.com] Im Auftrag von Brian Kirby >>> Gesendet: Donnerstag, 18. Juni 2009 01:51 >>> An: Discussion of precise time and frequency measurement >>> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up >>> >>> >>> They do not talk about driving the frequency off, but they >>> warn to keep >>> the receiver locked the first 24 hours, so it can determine how the >>> oscillator ages. Its in the section on "holdover" (page 52 >>> of the PDF , >>> page 3-8 of the user guide). >>> >>> You may also want to search for HP An1279, it also goes over HP >>> smartclock operation, and look for a paper titled "the Global >>> Positioning System and HP SmartClock" by John A. Kusters. >>> >>> If you have a broadband connection, I can email you each >>> >> one, if that >> >>> will help. >>> >>> Brian >>> >>> Ulrich Bangert wrote: >>> >>>> Brian, >>>> >>>> thanks for your information! >>>> >>>> >>>> >>>>> algorithm. From what I read about the smart clock, it >>>>> >> takes 5 days >> >>>>> for it to complete its initial learning cycle and then its >>>>> >>>>> >>>> continuously >>>> >>>> >>>>> refining. >>>>> >>>>> >>>> Is that from the Z3801 manual, which I have available, or >>>> >>> do you have >>> >>>> any other in depth information source? >>>> >>>> Best regards >>>> Ulrich >>>> >>>> >>>> >>>>> -----Ursprungliche Nachricht----- >>>>> Von: time-nuts-bounces@febo.com >>>>> >> [mailto:time-nuts-bounces@febo.com] >> >>>>> Im Auftrag von Brian Kirby >>>>> Gesendet: Mittwoch, 17. Juni 2009 16:58 >>>>> An: Discussion of precise time and frequency measurement >>>>> Betreff: Re: [time-nuts] Z3805 initial behaviour after power up >>>>> >>>>> >>>>> Ulrich, >>>>> >>>>> I have two Z3801A and both of them act close to what you >>>>> >> describe >> >>>>> in the first 24 hours of power up. I believe its part of the >>>>> >>> disciplining >>> >>>>> algorithm. From what I read about the smart clock, it takes >>>>> 5 days for >>>>> it to complete its initial learning cycle and then its >>>>> >>> continuously >>> >>>>> refining. >>>>> >>>>> Brian >>>>> >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@febo.com >>>>> To unsubscribe, go to >>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>>> and follow the instructions there. >>>>> >>>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com >>>> To unsubscribe, go to >>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>> and follow the instructions there. >>>> >>>> >>>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
PC
Paul Christensen
Fri, Jun 19, 2009 4:38 PM

"The other issue is noise from a switcher."

And be mindful that some frequency-standard devices contain integrated DC-DC
switching converters -- like my HP 58540A GPS-DO.  The noise from the HP DC
converters was spraying RFI every 100 kHz or so on the HF bands.  After
trying various RFI abatement measures, including additional bypass C and
toroid-wound #31 ferrite material, my solution was scrap the entire PS board
and bring in clean power from a triple-output linear supply made by
Power-One.  There's no switch-mode component anywhere in the path of my
GPS-DO.

I love switch-mode devices for their light weight and package density; I
hate them for their propensity to generate RFI and at least historically,
their abysmal long-term reliability.  That said, I do believe that recent
SMPS product is significantly more reliable than those manufactured 10+
years ago.

Paul, W9AC

> "The other issue is noise from a switcher." And be mindful that some frequency-standard devices contain integrated DC-DC switching converters -- like my HP 58540A GPS-DO. The noise from the HP DC converters was spraying RFI every 100 kHz or so on the HF bands. After trying various RFI abatement measures, including additional bypass C and toroid-wound #31 ferrite material, my solution was scrap the entire PS board and bring in clean power from a triple-output linear supply made by Power-One. There's no switch-mode component anywhere in the path of my GPS-DO. I love switch-mode devices for their light weight and package density; I hate them for their propensity to generate RFI and at least historically, their abysmal long-term reliability. That said, I do believe that recent SMPS product is significantly more reliable than those manufactured 10+ years ago. Paul, W9AC
BK
Brian Kirby
Sat, Jun 20, 2009 4:49 AM

I will have two Trimble Thunderbolts GPSDO, that I will be selling.

Both were bought thru the Time-Nuts purchases.  There is nothing wrong
with these units, they are operational, and surplus to my needs.

One unit comes with the Ault power supply and power connector.

The second unit comes without a power supply and I will be furnishing a
pig-tailed power lead connector.

I do not have pay-pal (sorry, I gave up on *bay...) and I can send COD.
I can only make this offer for the lower 48 of the United States.

If you are interested, please contact me off this list.

Brian Kirby - KD4FM

I will have two Trimble Thunderbolts GPSDO, that I will be selling. Both were bought thru the Time-Nuts purchases. There is nothing wrong with these units, they are operational, and surplus to my needs. One unit comes with the Ault power supply and power connector. The second unit comes without a power supply and I will be furnishing a pig-tailed power lead connector. I do not have pay-pal (sorry, I gave up on *bay...) and I can send COD. I can only make this offer for the lower 48 of the United States. If you are interested, please contact me off this list. Brian Kirby - KD4FM
MD
Magnus Danielson
Sat, Jun 20, 2009 11:33 AM

Francesco Ledda wrote:

Correction...

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On
Behalf Of Francesco Ledda
Sent: Friday, June 19, 2009 9:38 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Z3805 initial behavior after power up revised

Variation due to environment are deterministic and are nor part of aging.
Ageing needs to measured in conditions that remove ambient induced
perturbations.

Not entierly true, as aging behaviour may be re-started or modified as
environmental conditions change.

I used to put crystal oscillators in high quality
environmental chambers that kept temp, humidity and pressure constant and
measure the ageing.  There are tricks that can be done to increase the aging
rate, so that we did not have to wait 20 years ;)

Strange, as the aging rate decreases over time, most of its aging and
shift in frequency occur early. This occurs both in the "green" aging
period, when it is new, but also after it has been turned off and cooled
down / low temperature storage.

So what tricks did you use?

We know that ageing happens, but the direction of ageing cannot be
determined.
A coin toss can yield face or tail.  If the coin is perfect, the
distribution will be uniform.  If a face is 1 and a tail is -1, we can add
successive tosses and track the total number.  Even is the distribution of
tosses is uniform, the sum will walk away from 0 (random walk),cameback to 0
and then move away from 0 again.  This is a good way to model and explain
ageing. Nature is perfect, but things do not follow our math perfectly.  A
better simulation would include a "leaky integral".

The statistical analysis of ageing is tricky, since ageing doeas have NOT a
statistical mean, and therefore standard deviation cannot be used.

Standard deviation of frequency isn't used, but rather it is modeled
thru the frequency drift component, and this is also described for
different time-spans. Over time it integrates to the drifted frequency
over that time. In all models it is treated separately from the noise
components.

The curvature of aging has also been established (see for instance Vig
on the UFFC site), and with external reference (such as in a GPSDO), the
parameters for that curve can be established and adjusted over time for
better prediction of where the aging.

Or is it that we have different concepts or environments in mind when we
discuss aging? For me, aging is the shift in frequency over time as the
oscillator is maintained at the same temperature.

This discussion brings lots of good memories back.  The good old days when
SONET synchronization was a new thing, and we were creating new technology.

There are a few others here that was involved with that, and it would be
fun to have a few discussion on that.

Cheers,
Magnus

Francesco Ledda wrote: > Correction... > > -----Original Message----- > From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]On > Behalf Of Francesco Ledda > Sent: Friday, June 19, 2009 9:38 AM > To: Discussion of precise time and frequency measurement > Subject: Re: [time-nuts] Z3805 initial behavior after power up revised > > > Variation due to environment are deterministic and are nor part of aging. > Ageing needs to measured in conditions that remove ambient induced > perturbations. Not entierly true, as aging behaviour may be re-started or modified as environmental conditions change. > I used to put crystal oscillators in high quality > environmental chambers that kept temp, humidity and pressure constant and > measure the ageing. There are tricks that can be done to increase the aging > rate, so that we did not have to wait 20 years ;) > Strange, as the aging rate decreases over time, most of its aging and shift in frequency occur early. This occurs both in the "green" aging period, when it is new, but also after it has been turned off and cooled down / low temperature storage. So what tricks did you use? > We know that ageing happens, but the direction of ageing cannot be > determined. > A coin toss can yield face or tail. If the coin is perfect, the > distribution will be uniform. If a face is 1 and a tail is -1, we can add > successive tosses and track the total number. Even is the distribution of > tosses is uniform, the sum will walk away from 0 (random walk),cameback to 0 > and then move away from 0 again. This is a good way to model and explain > ageing. Nature is perfect, but things do not follow our math perfectly. A > better simulation would include a "leaky integral". > > The statistical analysis of ageing is tricky, since ageing doeas have NOT a > statistical mean, and therefore standard deviation cannot be used. > Standard deviation of frequency isn't used, but rather it is modeled thru the frequency drift component, and this is also described for different time-spans. Over time it integrates to the drifted frequency over that time. In all models it is treated separately from the noise components. The curvature of aging has also been established (see for instance Vig on the UFFC site), and with external reference (such as in a GPSDO), the parameters for that curve can be established and adjusted over time for better prediction of where the aging. Or is it that we have different concepts or environments in mind when we discuss aging? For me, aging is the shift in frequency over time as the oscillator is maintained at the same temperature. > This discussion brings lots of good memories back. The good old days when > SONET synchronization was a new thing, and we were creating new technology. > There are a few others here that was involved with that, and it would be fun to have a few discussion on that. Cheers, Magnus