Sorry guys if I am out of line here. I Just joined and I am not sure of
the mechanics yet.
I missed earlier messages in the thread. I gather someone is selling a
Symmetricom 5115 for $11000.  ?
I happen to have(for my use) a 5125 which is their latest and greatest. It
is similar to the others but takes a carrier up to 400 MHz.
It is a fantastic box in being easy to use and measuring a wide range of
phase noise and allan dev. It also has a very nice network based remote
interface.
I think we paid $30,000. for ours Unfortunately it (or the 5115) is out of
range for my home lab. It is fun and educational to play with.


Nicholas Wagner
Senior Engineering Specialist
Aerospace Corporation
(310) 336-2509
nicholas.s.wagner@aero.org


Sorry guys if I am out of line here. I Just joined and I am not sure of the mechanics yet. I missed earlier messages in the thread. I gather someone is selling a Symmetricom 5115 for $11000. ? I happen to have(for my use) a 5125 which is their latest and greatest. It is similar to the others but takes a carrier up to 400 MHz. It is a fantastic box in being easy to use and measuring a wide range of phase noise and allan dev. It also has a very nice network based remote interface. I think we paid $30,000. for ours Unfortunately it (or the 5115) is out of range for my home lab. It is fun and educational to play with. ================================= Nicholas Wagner Senior Engineering Specialist Aerospace Corporation (310) 336-2509 nicholas.s.wagner@aero.org =================================
JP
Jeffrey Pawlan
Wed, Apr 7, 2010 9:52 PM

Dear Nicholas,

Thank you for getting this discussion back to technical aspects related our
real quest for technical information.

Please look in your actual documentation for the current model 5125 and
confirm its actual specs. I had read some specs on the Symmetricom website
but I do not know for which model. Although automated and easy to use, I
remember that the phase noise floor was lackluster and makes me think that
this may be fine for very close-in measurments but not for the usual 100Hz
to 100KHz offset measurments. Please let us know what you find with this
equipment.

Regards,

Jeffrey Pawlan

Dear Nicholas, Thank you for getting this discussion back to technical aspects related our real quest for technical information. Please look in your actual documentation for the current model 5125 and confirm its actual specs. I had read some specs on the Symmetricom website but I do not know for which model. Although automated and easy to use, I remember that the phase noise floor was lackluster and makes me think that this may be fine for very close-in measurments but not for the usual 100Hz to 100KHz offset measurments. Please let us know what you find with this equipment. Regards, Jeffrey Pawlan
JM
John Miles
Wed, Apr 7, 2010 11:01 PM

Thank you for getting this discussion back to technical aspects
related our
real quest for technical information.

Please look in your actual documentation for the current model 5125 and
confirm its actual specs. I had read some specs on the
Symmetricom website
but I do not know for which model. Although automated and easy to use, I
remember that the phase noise floor was lackluster and makes me
think that
this may be fine for very close-in measurments but not for the
usual 100Hz
to 100KHz offset measurments. Please let us know what you find with this
equipment.

The 5115A is more useful as an ADEV test set than as a phase-noise test set.
Symmetricom can upgrade it to 5120A specs, which enables cross correlation
and improves the PN floor substantially, but I have no idea how much that
costs.

The 5125A is a 5120A with coverage extended to 400 MHz.

-- john, KE5FX

> Thank you for getting this discussion back to technical aspects > related our > real quest for technical information. > > Please look in your actual documentation for the current model 5125 and > confirm its actual specs. I had read some specs on the > Symmetricom website > but I do not know for which model. Although automated and easy to use, I > remember that the phase noise floor was lackluster and makes me > think that > this may be fine for very close-in measurments but not for the > usual 100Hz > to 100KHz offset measurments. Please let us know what you find with this > equipment. The 5115A is more useful as an ADEV test set than as a phase-noise test set. Symmetricom can upgrade it to 5120A specs, which enables cross correlation and improves the PN floor substantially, but I have no idea how much that costs. The 5125A is a 5120A with coverage extended to 400 MHz. -- john, KE5FX
JA
John Ackermann N8UR
Wed, Apr 7, 2010 11:04 PM

I just took a look at the datasheets for the 5115, 5120, and 5125.  .

The 5115 is the "base" model covering through 30 MHz and without a
cross-correlation capability.  At 10 MHz input frequency, its spec is
-133 dBc/Hz at 1 Hz offset, and -147 dBc/Hz at >100 Hz offset (not
spec'd at greater offsets, so presumably that's the floor).

The 5120 adds correlation to lower the noise floor.  At 10 MHz input
frequency, it is spec'd at <-145 dBc/Hz at 1 Hz, and <-175 dBc/Hz at 10 kHz.

The 5125 extends the 5120 frequency range to 400 MHz.  At 10 MHz its
"typical" performance is -145 dBc/Hz at 1 Hz offset and -170 dBc/Hz at
10 kHz (specified performance is about 5 dB worse; the other two
datasheets only specify spec, not typical).  The floor increases with
frequency, with the close-in noise going up more rapidly than the floor
further out.

All three are spec'd at 1 dB accuracy, and can measure at offsets down
to 1 milliHertz.  One thing that's nice about the TSC boxes compared to
traditional PN measurement systems is that the reference doesn't need to
be at the same frequency as the DUT.

John

Jeffrey Pawlan said the following on 04/07/2010 05:52 PM:

Dear Nicholas,

Thank you for getting this discussion back to technical aspects related
our real quest for technical information.

Please look in your actual documentation for the current model 5125 and
confirm its actual specs. I had read some specs on the Symmetricom
website but I do not know for which model. Although automated and easy
to use, I remember that the phase noise floor was lackluster and makes
me think that this may be fine for very close-in measurments but not for
the usual 100Hz to 100KHz offset measurments. Please let us know what
you find with this equipment.

Regards,

Jeffrey Pawlan


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

I just took a look at the datasheets for the 5115, 5120, and 5125. . The 5115 is the "base" model covering through 30 MHz and without a cross-correlation capability. At 10 MHz input frequency, its spec is -133 dBc/Hz at 1 Hz offset, and -147 dBc/Hz at >100 Hz offset (not spec'd at greater offsets, so presumably that's the floor). The 5120 adds correlation to lower the noise floor. At 10 MHz input frequency, it is spec'd at <-145 dBc/Hz at 1 Hz, and <-175 dBc/Hz at 10 kHz. The 5125 extends the 5120 frequency range to 400 MHz. At 10 MHz its "typical" performance is -145 dBc/Hz at 1 Hz offset and -170 dBc/Hz at 10 kHz (specified performance is about 5 dB worse; the other two datasheets only specify spec, not typical). The floor increases with frequency, with the close-in noise going up more rapidly than the floor further out. All three are spec'd at 1 dB accuracy, and can measure at offsets down to 1 milliHertz. One thing that's nice about the TSC boxes compared to traditional PN measurement systems is that the reference doesn't need to be at the same frequency as the DUT. John ---- Jeffrey Pawlan said the following on 04/07/2010 05:52 PM: > Dear Nicholas, > > Thank you for getting this discussion back to technical aspects related > our real quest for technical information. > > Please look in your actual documentation for the current model 5125 and > confirm its actual specs. I had read some specs on the Symmetricom > website but I do not know for which model. Although automated and easy > to use, I remember that the phase noise floor was lackluster and makes > me think that this may be fine for very close-in measurments but not for > the usual 100Hz to 100KHz offset measurments. Please let us know what > you find with this equipment. > > Regards, > > Jeffrey Pawlan > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BC
Bob Camp
Thu, Apr 8, 2010 12:14 AM

Hi

So basically what I get when I buy a 5115 is a limited frequency range box that isn't quite as good as (as in 40 db worse than floor) an HP 3048. Of course I'd need to buy a cheap synthesizer to go with the 3048 if I wanted to do the "any frequency" stuff at the 5115 level.

Bob

On Apr 7, 2010, at 7:04 PM, John Ackermann N8UR wrote:

I just took a look at the datasheets for the 5115, 5120, and 5125.  .

The 5115 is the "base" model covering through 30 MHz and without a cross-correlation capability.  At 10 MHz input frequency, its spec is -133 dBc/Hz at 1 Hz offset, and -147 dBc/Hz at >100 Hz offset (not spec'd at greater offsets, so presumably that's the floor).

The 5120 adds correlation to lower the noise floor.  At 10 MHz input frequency, it is spec'd at <-145 dBc/Hz at 1 Hz, and <-175 dBc/Hz at 10 kHz.

The 5125 extends the 5120 frequency range to 400 MHz.  At 10 MHz its "typical" performance is -145 dBc/Hz at 1 Hz offset and -170 dBc/Hz at 10 kHz (specified performance is about 5 dB worse; the other two datasheets only specify spec, not typical).  The floor increases with frequency, with the close-in noise going up more rapidly than the floor further out.

All three are spec'd at 1 dB accuracy, and can measure at offsets down to 1 milliHertz.  One thing that's nice about the TSC boxes compared to traditional PN measurement systems is that the reference doesn't need to be at the same frequency as the DUT.

John

Jeffrey Pawlan said the following on 04/07/2010 05:52 PM:

Dear Nicholas,
Thank you for getting this discussion back to technical aspects related our real quest for technical information.
Please look in your actual documentation for the current model 5125 and confirm its actual specs. I had read some specs on the Symmetricom website but I do not know for which model. Although automated and easy to use, I remember that the phase noise floor was lackluster and makes me think that this may be fine for very close-in measurments but not for the usual 100Hz to 100KHz offset measurments. Please let us know what you find with this equipment.
Regards,
Jeffrey Pawlan


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi So basically what I get when I buy a 5115 is a limited frequency range box that isn't quite as good as (as in 40 db worse than floor) an HP 3048. Of course I'd need to buy a cheap synthesizer to go with the 3048 if I wanted to do the "any frequency" stuff at the 5115 level. Bob On Apr 7, 2010, at 7:04 PM, John Ackermann N8UR wrote: > I just took a look at the datasheets for the 5115, 5120, and 5125. . > > The 5115 is the "base" model covering through 30 MHz and without a cross-correlation capability. At 10 MHz input frequency, its spec is -133 dBc/Hz at 1 Hz offset, and -147 dBc/Hz at >100 Hz offset (not spec'd at greater offsets, so presumably that's the floor). > > The 5120 adds correlation to lower the noise floor. At 10 MHz input frequency, it is spec'd at <-145 dBc/Hz at 1 Hz, and <-175 dBc/Hz at 10 kHz. > > The 5125 extends the 5120 frequency range to 400 MHz. At 10 MHz its "typical" performance is -145 dBc/Hz at 1 Hz offset and -170 dBc/Hz at 10 kHz (specified performance is about 5 dB worse; the other two datasheets only specify spec, not typical). The floor increases with frequency, with the close-in noise going up more rapidly than the floor further out. > > All three are spec'd at 1 dB accuracy, and can measure at offsets down to 1 milliHertz. One thing that's nice about the TSC boxes compared to traditional PN measurement systems is that the reference doesn't need to be at the same frequency as the DUT. > > John > ---- > > Jeffrey Pawlan said the following on 04/07/2010 05:52 PM: >> Dear Nicholas, >> Thank you for getting this discussion back to technical aspects related our real quest for technical information. >> Please look in your actual documentation for the current model 5125 and confirm its actual specs. I had read some specs on the Symmetricom website but I do not know for which model. Although automated and easy to use, I remember that the phase noise floor was lackluster and makes me think that this may be fine for very close-in measurments but not for the usual 100Hz to 100KHz offset measurments. Please let us know what you find with this equipment. >> Regards, >> Jeffrey Pawlan >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
JA
John Ackermann N8UR
Thu, Apr 8, 2010 12:25 PM

Well, you also get ADEV down to <1e-14 at tau = 1 second, and a 13 digit
frequency counter.  And as long as you're working in the HF range, the
big advantage over a PLL system is that you only need one reference
oscillator because the reference and DUT frequencies are independent.

As JohnM said, the 5115 is really more of an ADEV box.  As I understand
it, it was mainly intended to replace the 5110 which had very limited PN
capabilities.

John

Bob Camp said the following on 04/07/2010 08:14 PM:

Hi

So basically what I get when I buy a 5115 is a limited frequency range box that isn't quite as good as (as in 40 db worse than floor) an HP 3048. Of course I'd need to buy a cheap synthesizer to go with the 3048 if I wanted to do the "any frequency" stuff at the 5115 level.

Bob

On Apr 7, 2010, at 7:04 PM, John Ackermann N8UR wrote:

I just took a look at the datasheets for the 5115, 5120, and 5125.  .

The 5115 is the "base" model covering through 30 MHz and without a cross-correlation capability.  At 10 MHz input frequency, its spec is -133 dBc/Hz at 1 Hz offset, and -147 dBc/Hz at >100 Hz offset (not spec'd at greater offsets, so presumably that's the floor).

The 5120 adds correlation to lower the noise floor.  At 10 MHz input frequency, it is spec'd at <-145 dBc/Hz at 1 Hz, and <-175 dBc/Hz at 10 kHz.

The 5125 extends the 5120 frequency range to 400 MHz.  At 10 MHz its "typical" performance is -145 dBc/Hz at 1 Hz offset and -170 dBc/Hz at 10 kHz (specified performance is about 5 dB worse; the other two datasheets only specify spec, not typical).  The floor increases with frequency, with the close-in noise going up more rapidly than the floor further out.

All three are spec'd at 1 dB accuracy, and can measure at offsets down to 1 milliHertz.  One thing that's nice about the TSC boxes compared to traditional PN measurement systems is that the reference doesn't need to be at the same frequency as the DUT.

John

Jeffrey Pawlan said the following on 04/07/2010 05:52 PM:

Dear Nicholas,
Thank you for getting this discussion back to technical aspects related our real quest for technical information.
Please look in your actual documentation for the current model 5125 and confirm its actual specs. I had read some specs on the Symmetricom website but I do not know for which model. Although automated and easy to use, I remember that the phase noise floor was lackluster and makes me think that this may be fine for very close-in measurments but not for the usual 100Hz to 100KHz offset measurments. Please let us know what you find with this equipment.
Regards,
Jeffrey Pawlan


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Well, you also get ADEV down to <1e-14 at tau = 1 second, and a 13 digit frequency counter. And as long as you're working in the HF range, the big advantage over a PLL system is that you only need one reference oscillator because the reference and DUT frequencies are independent. As JohnM said, the 5115 is really more of an ADEV box. As I understand it, it was mainly intended to replace the 5110 which had very limited PN capabilities. John ---- Bob Camp said the following on 04/07/2010 08:14 PM: > Hi > > So basically what I get when I buy a 5115 is a limited frequency range box that isn't quite as good as (as in 40 db worse than floor) an HP 3048. Of course I'd need to buy a cheap synthesizer to go with the 3048 if I wanted to do the "any frequency" stuff at the 5115 level. > > Bob > > > On Apr 7, 2010, at 7:04 PM, John Ackermann N8UR wrote: > >> I just took a look at the datasheets for the 5115, 5120, and 5125. . >> >> The 5115 is the "base" model covering through 30 MHz and without a cross-correlation capability. At 10 MHz input frequency, its spec is -133 dBc/Hz at 1 Hz offset, and -147 dBc/Hz at >100 Hz offset (not spec'd at greater offsets, so presumably that's the floor). >> >> The 5120 adds correlation to lower the noise floor. At 10 MHz input frequency, it is spec'd at <-145 dBc/Hz at 1 Hz, and <-175 dBc/Hz at 10 kHz. >> >> The 5125 extends the 5120 frequency range to 400 MHz. At 10 MHz its "typical" performance is -145 dBc/Hz at 1 Hz offset and -170 dBc/Hz at 10 kHz (specified performance is about 5 dB worse; the other two datasheets only specify spec, not typical). The floor increases with frequency, with the close-in noise going up more rapidly than the floor further out. >> >> All three are spec'd at 1 dB accuracy, and can measure at offsets down to 1 milliHertz. One thing that's nice about the TSC boxes compared to traditional PN measurement systems is that the reference doesn't need to be at the same frequency as the DUT. >> >> John >> ---- >> >> Jeffrey Pawlan said the following on 04/07/2010 05:52 PM: >>> Dear Nicholas, >>> Thank you for getting this discussion back to technical aspects related our real quest for technical information. >>> Please look in your actual documentation for the current model 5125 and confirm its actual specs. I had read some specs on the Symmetricom website but I do not know for which model. Although automated and easy to use, I remember that the phase noise floor was lackluster and makes me think that this may be fine for very close-in measurments but not for the usual 100Hz to 100KHz offset measurments. Please let us know what you find with this equipment. >>> Regards, >>> Jeffrey Pawlan >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BC
Bob Camp
Thu, Apr 8, 2010 4:18 PM

Hi

True, the 3048 is not much of a frequency counter. It's also pretty limited
as an Adev box.

Bob

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of John Ackermann N8UR
Sent: Thursday, April 08, 2010 8:25 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Symmetricom 5115

Well, you also get ADEV down to <1e-14 at tau = 1 second, and a 13 digit
frequency counter.  And as long as you're working in the HF range, the
big advantage over a PLL system is that you only need one reference
oscillator because the reference and DUT frequencies are independent.

As JohnM said, the 5115 is really more of an ADEV box.  As I understand
it, it was mainly intended to replace the 5110 which had very limited PN
capabilities.

John

Bob Camp said the following on 04/07/2010 08:14 PM:

Hi

So basically what I get when I buy a 5115 is a limited frequency range box

that isn't quite as good as (as in 40 db worse than floor) an HP 3048. Of
course I'd need to buy a cheap synthesizer to go with the 3048 if I wanted
to do the "any frequency" stuff at the 5115 level.

Bob

On Apr 7, 2010, at 7:04 PM, John Ackermann N8UR wrote:

I just took a look at the datasheets for the 5115, 5120, and 5125.  .

The 5115 is the "base" model covering through 30 MHz and without a

cross-correlation capability.  At 10 MHz input frequency, its spec is -133
dBc/Hz at 1 Hz offset, and -147 dBc/Hz at >100 Hz offset (not spec'd at
greater offsets, so presumably that's the floor).

The 5120 adds correlation to lower the noise floor.  At 10 MHz input

frequency, it is spec'd at <-145 dBc/Hz at 1 Hz, and <-175 dBc/Hz at 10 kHz.

The 5125 extends the 5120 frequency range to 400 MHz.  At 10 MHz its

"typical" performance is -145 dBc/Hz at 1 Hz offset and -170 dBc/Hz at 10
kHz (specified performance is about 5 dB worse; the other two datasheets
only specify spec, not typical).  The floor increases with frequency, with
the close-in noise going up more rapidly than the floor further out.

All three are spec'd at 1 dB accuracy, and can measure at offsets down to

1 milliHertz.  One thing that's nice about the TSC boxes compared to
traditional PN measurement systems is that the reference doesn't need to be
at the same frequency as the DUT.

John

Jeffrey Pawlan said the following on 04/07/2010 05:52 PM:

Dear Nicholas,
Thank you for getting this discussion back to technical aspects related

our real quest for technical information.

Please look in your actual documentation for the current model 5125 and

confirm its actual specs. I had read some specs on the Symmetricom website
but I do not know for which model. Although automated and easy to use, I
remember that the phase noise floor was lackluster and makes me think that
this may be fine for very close-in measurments but not for the usual 100Hz
to 100KHz offset measurments. Please let us know what you find with this
equipment.

Regards,
Jeffrey Pawlan


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi True, the 3048 is not much of a frequency counter. It's also pretty limited as an Adev box. Bob -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On Behalf Of John Ackermann N8UR Sent: Thursday, April 08, 2010 8:25 AM To: Discussion of precise time and frequency measurement Subject: Re: [time-nuts] Symmetricom 5115 Well, you also get ADEV down to <1e-14 at tau = 1 second, and a 13 digit frequency counter. And as long as you're working in the HF range, the big advantage over a PLL system is that you only need one reference oscillator because the reference and DUT frequencies are independent. As JohnM said, the 5115 is really more of an ADEV box. As I understand it, it was mainly intended to replace the 5110 which had very limited PN capabilities. John ---- Bob Camp said the following on 04/07/2010 08:14 PM: > Hi > > So basically what I get when I buy a 5115 is a limited frequency range box that isn't quite as good as (as in 40 db worse than floor) an HP 3048. Of course I'd need to buy a cheap synthesizer to go with the 3048 if I wanted to do the "any frequency" stuff at the 5115 level. > > Bob > > > On Apr 7, 2010, at 7:04 PM, John Ackermann N8UR wrote: > >> I just took a look at the datasheets for the 5115, 5120, and 5125. . >> >> The 5115 is the "base" model covering through 30 MHz and without a cross-correlation capability. At 10 MHz input frequency, its spec is -133 dBc/Hz at 1 Hz offset, and -147 dBc/Hz at >100 Hz offset (not spec'd at greater offsets, so presumably that's the floor). >> >> The 5120 adds correlation to lower the noise floor. At 10 MHz input frequency, it is spec'd at <-145 dBc/Hz at 1 Hz, and <-175 dBc/Hz at 10 kHz. >> >> The 5125 extends the 5120 frequency range to 400 MHz. At 10 MHz its "typical" performance is -145 dBc/Hz at 1 Hz offset and -170 dBc/Hz at 10 kHz (specified performance is about 5 dB worse; the other two datasheets only specify spec, not typical). The floor increases with frequency, with the close-in noise going up more rapidly than the floor further out. >> >> All three are spec'd at 1 dB accuracy, and can measure at offsets down to 1 milliHertz. One thing that's nice about the TSC boxes compared to traditional PN measurement systems is that the reference doesn't need to be at the same frequency as the DUT. >> >> John >> ---- >> >> Jeffrey Pawlan said the following on 04/07/2010 05:52 PM: >>> Dear Nicholas, >>> Thank you for getting this discussion back to technical aspects related our real quest for technical information. >>> Please look in your actual documentation for the current model 5125 and confirm its actual specs. I had read some specs on the Symmetricom website but I do not know for which model. Although automated and easy to use, I remember that the phase noise floor was lackluster and makes me think that this may be fine for very close-in measurments but not for the usual 100Hz to 100KHz offset measurments. Please let us know what you find with this equipment. >>> Regards, >>> Jeffrey Pawlan >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
MD
Magnus Danielson
Thu, Apr 8, 2010 6:42 PM

Bob Camp wrote:

Hi

True, the 3048 is not much of a frequency counter. It's also pretty limited
as an Adev box.

Box???

Cheers,
Magnus

Bob Camp wrote: > Hi > > True, the 3048 is not much of a frequency counter. It's also pretty limited > as an Adev box. Box??? Cheers, Magnus
SR
Steve Rooke
Fri, Apr 9, 2010 2:26 AM

On 9 April 2010 06:42, Magnus Danielson magnus@rubidium.dyndns.org wrote:

Bob Camp wrote:

Hi

True, the 3048 is not much of a frequency counter. It's also pretty
limited
as an Adev box.

Box???

The portability kit comes complete with a fork lift.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
Steve Rooke - ZL3TUV & G8KVD
A man with one clock knows what time it is;
A man with two clocks is never quite sure.

On 9 April 2010 06:42, Magnus Danielson <magnus@rubidium.dyndns.org> wrote: > Bob Camp wrote: >> >> Hi >> >> True, the 3048 is not much of a frequency counter. It's also pretty >> limited >> as an Adev box. > > Box??? The portability kit comes complete with a fork lift. > Cheers, > Magnus > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > -- Steve Rooke - ZL3TUV & G8KVD A man with one clock knows what time it is; A man with two clocks is never quite sure.