JM
John Miller
Tue, Apr 4, 2023 2:27 AM

Hello all,

I recently got my hands on a "gpib4pi" board specifically so that I can
log measurements from my recently acquired Racal Dana 1992. I understand
that the 1992 has some oddities when it comes to GPIB, but based on what
I've read in the manual, and from others' experiences, I should be able
to do what I want with no problems. I have made sure to set the internal
jumper to "GPIB mode", and it is not in "talk only" mode per the switch
settings on the rear of the unit.

The gpib4pi board is interfaced with via the linux gpib tools and
libraries, and to start I'm using the ibterm software to test things out

  • I'll be using python for logging once I confirm everything is working
    correctly. The issue that I am running into is that while I can send
    commands with no problems (e.g. changing measurement mode, changing
    resolution, resetting the unit, changing trigger modes, etc.) I cannot
    seem to receive any data at all.

If I issue a command to have it send data, like GET, no data is
returned, and the unit's SRQ (system request) LED is lit until I issue
another command. I feel like this should be relatively trivial but I am
getting absolutely nowhere fiddling with settings.

I have also connected the gpib4pi to a 33120A function generator, and if
I issue the command *IDN? through ibterm I get the proper
self-identification string back, so I feel fairly confident that my
hardware setup is correct, there's just something unique about the 1992
that I haven't figured out yet.

Here's my gpib.conf file: https://paste.millerjs.org/oyugadohey.txt

Thanks!

John

KC1QLN

Hello all, I recently got my hands on a "gpib4pi" board specifically so that I can log measurements from my recently acquired Racal Dana 1992. I understand that the 1992 has some oddities when it comes to GPIB, but based on what I've read in the manual, and from others' experiences, I should be able to do what I want with no problems. I have made sure to set the internal jumper to "GPIB mode", and it is not in "talk only" mode per the switch settings on the rear of the unit. The gpib4pi board is interfaced with via the linux gpib tools and libraries, and to start I'm using the ibterm software to test things out - I'll be using python for logging once I confirm everything is working correctly. The issue that I am running into is that while I can send commands with no problems (e.g. changing measurement mode, changing resolution, resetting the unit, changing trigger modes, etc.) I cannot seem to receive any data at all. If I issue a command to have it send data, like GET, no data is returned, and the unit's SRQ (system request) LED is lit until I issue another command. I feel like this should be relatively trivial but I am getting absolutely nowhere fiddling with settings. I have also connected the gpib4pi to a 33120A function generator, and if I issue the command *IDN? through ibterm I get the proper self-identification string back, so I feel fairly confident that my hardware setup is correct, there's just something unique about the 1992 that I haven't figured out yet. Here's my gpib.conf file: https://paste.millerjs.org/oyugadohey.txt Thanks! John KC1QLN
AG
Adrian Godwin
Tue, Apr 4, 2023 10:42 PM

There's a group for Racal equipment : https://groups.io/g/RacalDana  - you
may be able to find specific info on the 1992 there. A recent thread
discussed using a Racal 1991 with the open source AR488 interface, which
emulates the Prologix serial adapter. The thread title is 'GPIB on a
budget?' but you have to be a member of the list to read it.

The poster in that case did end up putting the counter in talk-only mode in
order to get regular updates out of it. His 1991 used an older pre-SCPI
version of GPIB but it should still be possible to communicate. However
automatic operations such as going into read mode after a '?' won't help. I
don't know if the 1992 is similar.

It might be helpful to connect a monitor to the bus to determine what state
it's in when it stalls. The sigrok project can be used to create a simple
logic analyser from a cheap Cypress Fx2 dev/breakout and then analyse it
with a postprocessor that disassembles the GPIB bus protocol.

On Tue, Apr 4, 2023 at 8:44 PM John Miller via time-nuts <
time-nuts@lists.febo.com> wrote:

Hello all,

I recently got my hands on a "gpib4pi" board specifically so that I can
log measurements from my recently acquired Racal Dana 1992. I understand
that the 1992 has some oddities when it comes to GPIB, but based on what
I've read in the manual, and from others' experiences, I should be able
to do what I want with no problems. I have made sure to set the internal
jumper to "GPIB mode", and it is not in "talk only" mode per the switch
settings on the rear of the unit.

The gpib4pi board is interfaced with via the linux gpib tools and
libraries, and to start I'm using the ibterm software to test things out

  • I'll be using python for logging once I confirm everything is working
    correctly. The issue that I am running into is that while I can send
    commands with no problems (e.g. changing measurement mode, changing
    resolution, resetting the unit, changing trigger modes, etc.) I cannot
    seem to receive any data at all.

If I issue a command to have it send data, like GET, no data is
returned, and the unit's SRQ (system request) LED is lit until I issue
another command. I feel like this should be relatively trivial but I am
getting absolutely nowhere fiddling with settings.

I have also connected the gpib4pi to a 33120A function generator, and if
I issue the command *IDN? through ibterm I get the proper
self-identification string back, so I feel fairly confident that my
hardware setup is correct, there's just something unique about the 1992
that I haven't figured out yet.

Here's my gpib.conf file: https://paste.millerjs.org/oyugadohey.txt

Thanks!

John

KC1QLN


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

There's a group for Racal equipment : https://groups.io/g/RacalDana - you may be able to find specific info on the 1992 there. A recent thread discussed using a Racal 1991 with the open source AR488 interface, which emulates the Prologix serial adapter. The thread title is 'GPIB on a budget?' but you have to be a member of the list to read it. The poster in that case did end up putting the counter in talk-only mode in order to get regular updates out of it. His 1991 used an older pre-SCPI version of GPIB but it should still be possible to communicate. However automatic operations such as going into read mode after a '?' won't help. I don't know if the 1992 is similar. It might be helpful to connect a monitor to the bus to determine what state it's in when it stalls. The sigrok project can be used to create a simple logic analyser from a cheap Cypress Fx2 dev/breakout and then analyse it with a postprocessor that disassembles the GPIB bus protocol. On Tue, Apr 4, 2023 at 8:44 PM John Miller via time-nuts < time-nuts@lists.febo.com> wrote: > Hello all, > > > I recently got my hands on a "gpib4pi" board specifically so that I can > log measurements from my recently acquired Racal Dana 1992. I understand > that the 1992 has some oddities when it comes to GPIB, but based on what > I've read in the manual, and from others' experiences, I should be able > to do what I want with no problems. I have made sure to set the internal > jumper to "GPIB mode", and it is not in "talk only" mode per the switch > settings on the rear of the unit. > > > The gpib4pi board is interfaced with via the linux gpib tools and > libraries, and to start I'm using the ibterm software to test things out > - I'll be using python for logging once I confirm everything is working > correctly. The issue that I am running into is that while I can send > commands with no problems (e.g. changing measurement mode, changing > resolution, resetting the unit, changing trigger modes, etc.) I cannot > seem to receive any data at all. > > > If I issue a command to have it send data, like GET, no data is > returned, and the unit's SRQ (system request) LED is lit until I issue > another command. I feel like this should be relatively trivial but I am > getting absolutely nowhere fiddling with settings. > > > I have also connected the gpib4pi to a 33120A function generator, and if > I issue the command *IDN? through ibterm I get the proper > self-identification string back, so I feel fairly confident that my > hardware setup is correct, there's just something unique about the 1992 > that I haven't figured out yet. > > > Here's my gpib.conf file: https://paste.millerjs.org/oyugadohey.txt > > > Thanks! > > John > > KC1QLN > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com >
JM
John Miller
Tue, Apr 4, 2023 11:07 PM

Hi Adrian,
Thanks! This is exactly what I was looking for - I had poked around for a Racal Dana group but wasn't able to find one, go figure. I've submitted a membership request and hopefully I can gain access soon. I'm open to doing bus analysis but I want to exhaust more options before I go that route. I'm not against running it in talk-only mode, either, I just want it to run long term data collection tasks - I don't need it to be fully remotely controllable.

John
KC1QLN

On Apr 4, 2023, at 6:42 PM, Adrian Godwin artgodwin@gmail.com wrote:

There's a group for Racal equipment : https://groups.io/g/RacalDana  - you may be able to find specific info on the 1992 there. A recent thread discussed using a Racal 1991 with the open source AR488 interface, which emulates the Prologix serial adapter. The thread title is 'GPIB on a budget?' but you have to be a member of the list to read it.

The poster in that case did end up putting the counter in talk-only mode in order to get regular updates out of it. His 1991 used an older pre-SCPI version of GPIB but it should still be possible to communicate. However automatic operations such as going into read mode after a '?' won't help. I don't know if the 1992 is similar.

It might be helpful to connect a monitor to the bus to determine what state it's in when it stalls. The sigrok project can be used to create a simple logic analyser from a cheap Cypress Fx2 dev/breakout and then analyse it with a postprocessor that disassembles the GPIB bus protocol.

On Tue, Apr 4, 2023 at 8:44 PM John Miller via time-nuts <time-nuts@lists.febo.com mailto:time-nuts@lists.febo.com> wrote:

Hello all,

I recently got my hands on a "gpib4pi" board specifically so that I can
log measurements from my recently acquired Racal Dana 1992. I understand
that the 1992 has some oddities when it comes to GPIB, but based on what
I've read in the manual, and from others' experiences, I should be able
to do what I want with no problems. I have made sure to set the internal
jumper to "GPIB mode", and it is not in "talk only" mode per the switch
settings on the rear of the unit.

The gpib4pi board is interfaced with via the linux gpib tools and
libraries, and to start I'm using the ibterm software to test things out

  • I'll be using python for logging once I confirm everything is working
    correctly. The issue that I am running into is that while I can send
    commands with no problems (e.g. changing measurement mode, changing
    resolution, resetting the unit, changing trigger modes, etc.) I cannot
    seem to receive any data at all.

If I issue a command to have it send data, like GET, no data is
returned, and the unit's SRQ (system request) LED is lit until I issue
another command. I feel like this should be relatively trivial but I am
getting absolutely nowhere fiddling with settings.

I have also connected the gpib4pi to a 33120A function generator, and if
I issue the command *IDN? through ibterm I get the proper
self-identification string back, so I feel fairly confident that my
hardware setup is correct, there's just something unique about the 1992
that I haven't figured out yet.

Here's my gpib.conf file: https://paste.millerjs.org/oyugadohey.txt

Thanks!

John

KC1QLN


time-nuts mailing list -- time-nuts@lists.febo.com mailto:time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com mailto:time-nuts-leave@lists.febo.com

Hi Adrian, Thanks! This is exactly what I was looking for - I had poked around for a Racal Dana group but wasn't able to find one, go figure. I've submitted a membership request and hopefully I can gain access soon. I'm open to doing bus analysis but I want to exhaust more options before I go that route. I'm not against running it in talk-only mode, either, I just want it to run long term data collection tasks - I don't need it to be fully remotely controllable. John KC1QLN > On Apr 4, 2023, at 6:42 PM, Adrian Godwin <artgodwin@gmail.com> wrote: > > There's a group for Racal equipment : https://groups.io/g/RacalDana - you may be able to find specific info on the 1992 there. A recent thread discussed using a Racal 1991 with the open source AR488 interface, which emulates the Prologix serial adapter. The thread title is 'GPIB on a budget?' but you have to be a member of the list to read it. > > The poster in that case did end up putting the counter in talk-only mode in order to get regular updates out of it. His 1991 used an older pre-SCPI version of GPIB but it should still be possible to communicate. However automatic operations such as going into read mode after a '?' won't help. I don't know if the 1992 is similar. > > It might be helpful to connect a monitor to the bus to determine what state it's in when it stalls. The sigrok project can be used to create a simple logic analyser from a cheap Cypress Fx2 dev/breakout and then analyse it with a postprocessor that disassembles the GPIB bus protocol. > > > On Tue, Apr 4, 2023 at 8:44 PM John Miller via time-nuts <time-nuts@lists.febo.com <mailto:time-nuts@lists.febo.com>> wrote: >> Hello all, >> >> >> I recently got my hands on a "gpib4pi" board specifically so that I can >> log measurements from my recently acquired Racal Dana 1992. I understand >> that the 1992 has some oddities when it comes to GPIB, but based on what >> I've read in the manual, and from others' experiences, I should be able >> to do what I want with no problems. I have made sure to set the internal >> jumper to "GPIB mode", and it is not in "talk only" mode per the switch >> settings on the rear of the unit. >> >> >> The gpib4pi board is interfaced with via the linux gpib tools and >> libraries, and to start I'm using the ibterm software to test things out >> - I'll be using python for logging once I confirm everything is working >> correctly. The issue that I am running into is that while I can send >> commands with no problems (e.g. changing measurement mode, changing >> resolution, resetting the unit, changing trigger modes, etc.) I cannot >> seem to receive any data at all. >> >> >> If I issue a command to have it send data, like GET, no data is >> returned, and the unit's SRQ (system request) LED is lit until I issue >> another command. I feel like this should be relatively trivial but I am >> getting absolutely nowhere fiddling with settings. >> >> >> I have also connected the gpib4pi to a 33120A function generator, and if >> I issue the command *IDN? through ibterm I get the proper >> self-identification string back, so I feel fairly confident that my >> hardware setup is correct, there's just something unique about the 1992 >> that I haven't figured out yet. >> >> >> Here's my gpib.conf file: https://paste.millerjs.org/oyugadohey.txt >> >> >> Thanks! >> >> John >> >> KC1QLN >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com <mailto:time-nuts@lists.febo.com> >> To unsubscribe send an email to time-nuts-leave@lists.febo.com <mailto:time-nuts-leave@lists.febo.com>
AG
Adrian Godwin
Wed, Apr 5, 2023 11:04 PM

The service manual for the 1991 (I assume the 1992 is similar) describes
how to check out the GPIB interface using a series of examples conducted on
an HP85. It doesn't mention any unusual settings but the process followed
may result in the HP85 asserting particular controls : not all setups will
do the same, as the GPIB commands contain some redundancy.

However, the HP85 is a good example even if you don't have one, as it was
used as an instrument controller for years. If you need details of what it
does on the bus, ask in another group, https://groups.io/g/hpseries80 which
is dedicated to the HP85 and related systems.

On Wed, Apr 5, 2023 at 12:14 AM John Miller via time-nuts <
time-nuts@lists.febo.com> wrote:

Hi Adrian,
Thanks! This is exactly what I was looking for - I had poked around for a
Racal Dana group but wasn't able to find one, go figure. I've submitted a
membership request and hopefully I can gain access soon. I'm open to doing
bus analysis but I want to exhaust more options before I go that route. I'm
not against running it in talk-only mode, either, I just want it to run
long term data collection tasks - I don't need it to be fully remotely
controllable.

John
KC1QLN

On Apr 4, 2023, at 6:42 PM, Adrian Godwin artgodwin@gmail.com wrote:

There's a group for Racal equipment : https://groups.io/g/RacalDana  -

you may be able to find specific info on the 1992 there. A recent thread
discussed using a Racal 1991 with the open source AR488 interface, which
emulates the Prologix serial adapter. The thread title is 'GPIB on a
budget?' but you have to be a member of the list to read it.

The poster in that case did end up putting the counter in talk-only mode

in order to get regular updates out of it. His 1991 used an older pre-SCPI
version of GPIB but it should still be possible to communicate. However
automatic operations such as going into read mode after a '?' won't help. I
don't know if the 1992 is similar.

It might be helpful to connect a monitor to the bus to determine what

state it's in when it stalls. The sigrok project can be used to create a
simple logic analyser from a cheap Cypress Fx2 dev/breakout and then
analyse it with a postprocessor that disassembles the GPIB bus protocol.

On Tue, Apr 4, 2023 at 8:44 PM John Miller via time-nuts <

Hello all,

I recently got my hands on a "gpib4pi" board specifically so that I can
log measurements from my recently acquired Racal Dana 1992. I

understand

that the 1992 has some oddities when it comes to GPIB, but based on

what

I've read in the manual, and from others' experiences, I should be able
to do what I want with no problems. I have made sure to set the

internal

jumper to "GPIB mode", and it is not in "talk only" mode per the switch
settings on the rear of the unit.

The gpib4pi board is interfaced with via the linux gpib tools and
libraries, and to start I'm using the ibterm software to test things

out

  • I'll be using python for logging once I confirm everything is working
    correctly. The issue that I am running into is that while I can send
    commands with no problems (e.g. changing measurement mode, changing
    resolution, resetting the unit, changing trigger modes, etc.) I cannot
    seem to receive any data at all.

If I issue a command to have it send data, like GET, no data is
returned, and the unit's SRQ (system request) LED is lit until I issue
another command. I feel like this should be relatively trivial but I am
getting absolutely nowhere fiddling with settings.

I have also connected the gpib4pi to a 33120A function generator, and

if

I issue the command *IDN? through ibterm I get the proper
self-identification string back, so I feel fairly confident that my
hardware setup is correct, there's just something unique about the 1992
that I haven't figured out yet.

Here's my gpib.conf file: https://paste.millerjs.org/oyugadohey.txt

Thanks!

John

KC1QLN


time-nuts mailing list -- time-nuts@lists.febo.com <mailto:

To unsubscribe send an email to time-nuts-leave@lists.febo.com <mailto:

time-nuts-leave@lists.febo.com>


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

The service manual for the 1991 (I assume the 1992 is similar) describes how to check out the GPIB interface using a series of examples conducted on an HP85. It doesn't mention any unusual settings but the process followed may result in the HP85 asserting particular controls : not all setups will do the same, as the GPIB commands contain some redundancy. However, the HP85 is a good example even if you don't have one, as it was used as an instrument controller for years. If you need details of what it does on the bus, ask in another group, https://groups.io/g/hpseries80 which is dedicated to the HP85 and related systems. On Wed, Apr 5, 2023 at 12:14 AM John Miller via time-nuts < time-nuts@lists.febo.com> wrote: > Hi Adrian, > Thanks! This is exactly what I was looking for - I had poked around for a > Racal Dana group but wasn't able to find one, go figure. I've submitted a > membership request and hopefully I can gain access soon. I'm open to doing > bus analysis but I want to exhaust more options before I go that route. I'm > not against running it in talk-only mode, either, I just want it to run > long term data collection tasks - I don't need it to be fully remotely > controllable. > > John > KC1QLN > > > On Apr 4, 2023, at 6:42 PM, Adrian Godwin <artgodwin@gmail.com> wrote: > > > > There's a group for Racal equipment : https://groups.io/g/RacalDana - > you may be able to find specific info on the 1992 there. A recent thread > discussed using a Racal 1991 with the open source AR488 interface, which > emulates the Prologix serial adapter. The thread title is 'GPIB on a > budget?' but you have to be a member of the list to read it. > > > > The poster in that case did end up putting the counter in talk-only mode > in order to get regular updates out of it. His 1991 used an older pre-SCPI > version of GPIB but it should still be possible to communicate. However > automatic operations such as going into read mode after a '?' won't help. I > don't know if the 1992 is similar. > > > > It might be helpful to connect a monitor to the bus to determine what > state it's in when it stalls. The sigrok project can be used to create a > simple logic analyser from a cheap Cypress Fx2 dev/breakout and then > analyse it with a postprocessor that disassembles the GPIB bus protocol. > > > > > > On Tue, Apr 4, 2023 at 8:44 PM John Miller via time-nuts < > time-nuts@lists.febo.com <mailto:time-nuts@lists.febo.com>> wrote: > >> Hello all, > >> > >> > >> I recently got my hands on a "gpib4pi" board specifically so that I can > >> log measurements from my recently acquired Racal Dana 1992. I > understand > >> that the 1992 has some oddities when it comes to GPIB, but based on > what > >> I've read in the manual, and from others' experiences, I should be able > >> to do what I want with no problems. I have made sure to set the > internal > >> jumper to "GPIB mode", and it is not in "talk only" mode per the switch > >> settings on the rear of the unit. > >> > >> > >> The gpib4pi board is interfaced with via the linux gpib tools and > >> libraries, and to start I'm using the ibterm software to test things > out > >> - I'll be using python for logging once I confirm everything is working > >> correctly. The issue that I am running into is that while I can send > >> commands with no problems (e.g. changing measurement mode, changing > >> resolution, resetting the unit, changing trigger modes, etc.) I cannot > >> seem to receive any data at all. > >> > >> > >> If I issue a command to have it send data, like GET, no data is > >> returned, and the unit's SRQ (system request) LED is lit until I issue > >> another command. I feel like this should be relatively trivial but I am > >> getting absolutely nowhere fiddling with settings. > >> > >> > >> I have also connected the gpib4pi to a 33120A function generator, and > if > >> I issue the command *IDN? through ibterm I get the proper > >> self-identification string back, so I feel fairly confident that my > >> hardware setup is correct, there's just something unique about the 1992 > >> that I haven't figured out yet. > >> > >> > >> Here's my gpib.conf file: https://paste.millerjs.org/oyugadohey.txt > >> > >> > >> Thanks! > >> > >> John > >> > >> KC1QLN > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@lists.febo.com <mailto: > time-nuts@lists.febo.com> > >> To unsubscribe send an email to time-nuts-leave@lists.febo.com <mailto: > time-nuts-leave@lists.febo.com> > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com