JG
Joseph Gray
Mon, Oct 5, 2009 7:08 AM
Sorry for the newbie question, but with two Z3801A's, what use would I
find for a rubidium unit? I see lots of them on ebay from the Chinese
sellers.
Sorry for the newbie question, but with two Z3801A's, what use would I
find for a rubidium unit? I see lots of them on ebay from the Chinese
sellers.
JL
J. L. Trantham
Mon, Oct 5, 2009 12:36 PM
Rubidium is a 'Secondary' reference, meaning it must be calibrated against a
'Primary' reference such as a Cesium Beam or Hydrogen Maser or, perhaps,
your GPSDO. Cesium Beam Standards some times appear on e..y in various
states of disrepair but are fun to play with. There have been many
discussions about their merits, or lack there of, on this list and it might
make interesting reading. I am amazed that, in my little shop, I can
generate something that seems 'locked' to the signal I get from my GPSDO and
requires no calibration.
Your GPSDO is 'linked' to the NIST Hydrogen Maser via the satellites. The
only problem comes when and if the satellites are not usable.
I am a relative 'newbie' as well, so be careful. This stuff is addicting.
Joe
-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Joseph Gray
Sent: Monday, October 05, 2009 2:09 AM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] Why would I want a rubidium
Sorry for the newbie question, but with two Z3801A's, what use would I find
for a rubidium unit? I see lots of them on ebay from the Chinese sellers.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Rubidium is a 'Secondary' reference, meaning it must be calibrated against a
'Primary' reference such as a Cesium Beam or Hydrogen Maser or, perhaps,
your GPSDO. Cesium Beam Standards some times appear on e..y in various
states of disrepair but are fun to play with. There have been many
discussions about their merits, or lack there of, on this list and it might
make interesting reading. I am amazed that, in my little shop, I can
generate something that seems 'locked' to the signal I get from my GPSDO and
requires no calibration.
Your GPSDO is 'linked' to the NIST Hydrogen Maser via the satellites. The
only problem comes when and if the satellites are not usable.
I am a relative 'newbie' as well, so be careful. This stuff is addicting.
Joe
-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Joseph Gray
Sent: Monday, October 05, 2009 2:09 AM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] Why would I want a rubidium
Sorry for the newbie question, but with two Z3801A's, what use would I find
for a rubidium unit? I see lots of them on ebay from the Chinese sellers.
_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
JF
J. Forster
Mon, Oct 5, 2009 4:58 PM
From an engineering point of view:
Very few people really "NEED" a Rb or Cs as they are not really doing
anything that requires that level of precision in frequency or accuracy in
time (me included). These days a BC-221 or WWB is not good enough, but 1
Hz at X-Band is for almost all practical uses.
It's a hobby. Just enjoy it!
FWIW,
-John
================
Rubidium is a 'Secondary' reference, meaning it must be calibrated against
a
'Primary' reference such as a Cesium Beam or Hydrogen Maser or, perhaps,
your GPSDO. Cesium Beam Standards some times appear on e..y in various
states of disrepair but are fun to play with. There have been many
discussions about their merits, or lack there of, on this list and it
might
make interesting reading. I am amazed that, in my little shop, I can
generate something that seems 'locked' to the signal I get from my GPSDO
and
requires no calibration.
Your GPSDO is 'linked' to the NIST Hydrogen Maser via the satellites. The
only problem comes when and if the satellites are not usable.
I am a relative 'newbie' as well, so be careful. This stuff is addicting.
Joe
-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Joseph Gray
Sent: Monday, October 05, 2009 2:09 AM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] Why would I want a rubidium
Sorry for the newbie question, but with two Z3801A's, what use would I
find
for a rubidium unit? I see lots of them on ebay from the Chinese sellers.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
>From an engineering point of view:
Very few people really "NEED" a Rb or Cs as they are not really doing
anything that requires that level of precision in frequency or accuracy in
time (me included). These days a BC-221 or WWB is not good enough, but 1
Hz at X-Band is for almost all practical uses.
It's a hobby. Just enjoy it!
FWIW,
-John
================
> Rubidium is a 'Secondary' reference, meaning it must be calibrated against
> a
> 'Primary' reference such as a Cesium Beam or Hydrogen Maser or, perhaps,
> your GPSDO. Cesium Beam Standards some times appear on e..y in various
> states of disrepair but are fun to play with. There have been many
> discussions about their merits, or lack there of, on this list and it
> might
> make interesting reading. I am amazed that, in my little shop, I can
> generate something that seems 'locked' to the signal I get from my GPSDO
> and
> requires no calibration.
>
> Your GPSDO is 'linked' to the NIST Hydrogen Maser via the satellites. The
> only problem comes when and if the satellites are not usable.
>
> I am a relative 'newbie' as well, so be careful. This stuff is addicting.
>
> Joe
>
>
>
> -----Original Message-----
> From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
> Behalf Of Joseph Gray
> Sent: Monday, October 05, 2009 2:09 AM
> To: Discussion of precise time and frequency measurement
> Subject: [time-nuts] Why would I want a rubidium
>
>
> Sorry for the newbie question, but with two Z3801A's, what use would I
> find
> for a rubidium unit? I see lots of them on ebay from the Chinese sellers.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
JG
Joseph Gray
Mon, Oct 5, 2009 5:20 PM
I know I don't "need" any of this stuff. I was just wondering what I
could do with a rubidium vs what I already have.
On Mon, Oct 5, 2009 at 10:58 AM, J. Forster jfor@quik.com wrote:
From an engineering point of view:
Very few people really "NEED" a Rb or Cs as they are not really doing
anything that requires that level of precision in frequency or accuracy in
time (me included). These days a BC-221 or WWB is not good enough, but 1
Hz at X-Band is for almost all practical uses.
It's a hobby. Just enjoy it!
FWIW,
-John
================
Rubidium is a 'Secondary' reference, meaning it must be calibrated against
a
'Primary' reference such as a Cesium Beam or Hydrogen Maser or, perhaps,
your GPSDO. Cesium Beam Standards some times appear on e..y in various
states of disrepair but are fun to play with. There have been many
discussions about their merits, or lack there of, on this list and it
might
make interesting reading. I am amazed that, in my little shop, I can
generate something that seems 'locked' to the signal I get from my GPSDO
and
requires no calibration.
Your GPSDO is 'linked' to the NIST Hydrogen Maser via the satellites. The
only problem comes when and if the satellites are not usable.
I am a relative 'newbie' as well, so be careful. This stuff is addicting.
Joe
-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Joseph Gray
Sent: Monday, October 05, 2009 2:09 AM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] Why would I want a rubidium
Sorry for the newbie question, but with two Z3801A's, what use would I
find
for a rubidium unit? I see lots of them on ebay from the Chinese sellers.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
I know I don't "need" any of this stuff. I was just wondering what I
could do with a rubidium vs what I already have.
On Mon, Oct 5, 2009 at 10:58 AM, J. Forster <jfor@quik.com> wrote:
> From an engineering point of view:
>
> Very few people really "NEED" a Rb or Cs as they are not really doing
> anything that requires that level of precision in frequency or accuracy in
> time (me included). These days a BC-221 or WWB is not good enough, but 1
> Hz at X-Band is for almost all practical uses.
>
> It's a hobby. Just enjoy it!
>
> FWIW,
> -John
>
> ================
>
>
>> Rubidium is a 'Secondary' reference, meaning it must be calibrated against
>> a
>> 'Primary' reference such as a Cesium Beam or Hydrogen Maser or, perhaps,
>> your GPSDO. Cesium Beam Standards some times appear on e..y in various
>> states of disrepair but are fun to play with. There have been many
>> discussions about their merits, or lack there of, on this list and it
>> might
>> make interesting reading. I am amazed that, in my little shop, I can
>> generate something that seems 'locked' to the signal I get from my GPSDO
>> and
>> requires no calibration.
>>
>> Your GPSDO is 'linked' to the NIST Hydrogen Maser via the satellites. The
>> only problem comes when and if the satellites are not usable.
>>
>> I am a relative 'newbie' as well, so be careful. This stuff is addicting.
>>
>> Joe
>>
>>
>>
>> -----Original Message-----
>> From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
>> Behalf Of Joseph Gray
>> Sent: Monday, October 05, 2009 2:09 AM
>> To: Discussion of precise time and frequency measurement
>> Subject: [time-nuts] Why would I want a rubidium
>>
>>
>> Sorry for the newbie question, but with two Z3801A's, what use would I
>> find
>> for a rubidium unit? I see lots of them on ebay from the Chinese sellers.
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>>
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
JA
John Ackermann N8UR
Mon, Oct 5, 2009 5:34 PM
Here's a reason why you might "want" (as opposed to "need"!) another
standard beyond the GPSDO.
An Rb or Cs is self-contained. The GPSDO relies on an external factor,
the presence of GPS. Holdover mode may be good to keep thing close for
a while, but over the long term the GPSDO without GPS is no better than
its crystal oscillator.
The Rb is a secondary standard and therefore isn't "correct" by
definition, but it has a low aging rate and gives you something
independent of GPS to use for measurements. Monitor it against GPS for
a while to learn its offset and drift characteristics, and then you can
extrapolate its performance out over a much longer time than you could
with an OCXO.
You may not require that independence, but it gives you additional
measurement capability. For example, comparing the output of two GPSDO
may not be meaningful because their frequencies could be correlated by
their common view of the GPS constellation. Using an Rb reference would
eliminate that common mode error and reveal information about the
GPSDO's short and medium term stability that would otherwise be hidden.
John
Joseph Gray wrote:
I know I don't "need" any of this stuff. I was just wondering what I
could do with a rubidium vs what I already have.
On Mon, Oct 5, 2009 at 10:58 AM, J. Forster jfor@quik.com wrote:
From an engineering point of view:
Very few people really "NEED" a Rb or Cs as they are not really doing
anything that requires that level of precision in frequency or accuracy in
time (me included). These days a BC-221 or WWB is not good enough, but 1
Hz at X-Band is for almost all practical uses.
Here's a reason why you might "want" (as opposed to "need"!) another
standard beyond the GPSDO.
An Rb or Cs is self-contained. The GPSDO relies on an external factor,
the presence of GPS. Holdover mode may be good to keep thing close for
a while, but over the long term the GPSDO without GPS is no better than
its crystal oscillator.
The Rb is a secondary standard and therefore isn't "correct" by
definition, but it has a low aging rate and gives you something
independent of GPS to use for measurements. Monitor it against GPS for
a while to learn its offset and drift characteristics, and then you can
extrapolate its performance out over a much longer time than you could
with an OCXO.
You may not require that independence, but it gives you additional
measurement capability. For example, comparing the output of two GPSDO
may not be meaningful because their frequencies could be correlated by
their common view of the GPS constellation. Using an Rb reference would
eliminate that common mode error and reveal information about the
GPSDO's short and medium term stability that would otherwise be hidden.
John
----
Joseph Gray wrote:
> I know I don't "need" any of this stuff. I was just wondering what I
> could do with a rubidium vs what I already have.
>
> On Mon, Oct 5, 2009 at 10:58 AM, J. Forster <jfor@quik.com> wrote:
>> From an engineering point of view:
>>
>> Very few people really "NEED" a Rb or Cs as they are not really doing
>> anything that requires that level of precision in frequency or accuracy in
>> time (me included). These days a BC-221 or WWB is not good enough, but 1
>> Hz at X-Band is for almost all practical uses.
JG
Joseph Gray
Mon, Oct 5, 2009 6:02 PM
Thanks, John. That was the type of information I was looking for.
Joe Gray
KA5ZEC
On Mon, Oct 5, 2009 at 11:34 AM, John Ackermann N8UR jra@febo.com wrote:
Here's a reason why you might "want" (as opposed to "need"!) another
standard beyond the GPSDO.
An Rb or Cs is self-contained. The GPSDO relies on an external factor, the
presence of GPS. Holdover mode may be good to keep thing close for a while,
but over the long term the GPSDO without GPS is no better than its crystal
oscillator.
The Rb is a secondary standard and therefore isn't "correct" by definition,
but it has a low aging rate and gives you something independent of GPS to
use for measurements. Monitor it against GPS for a while to learn its
offset and drift characteristics, and then you can extrapolate its
performance out over a much longer time than you could with an OCXO.
You may not require that independence, but it gives you additional
measurement capability. For example, comparing the output of two GPSDO may
not be meaningful because their frequencies could be correlated by their
common view of the GPS constellation. Using an Rb reference would eliminate
that common mode error and reveal information about the GPSDO's short and
medium term stability that would otherwise be hidden.
John
Joseph Gray wrote:
I know I don't "need" any of this stuff. I was just wondering what I
could do with a rubidium vs what I already have.
On Mon, Oct 5, 2009 at 10:58 AM, J. Forster jfor@quik.com wrote:
From an engineering point of view:
Very few people really "NEED" a Rb or Cs as they are not really doing
anything that requires that level of precision in frequency or accuracy
in
time (me included). These days a BC-221 or WWB is not good enough, but 1
Hz at X-Band is for almost all practical uses.
Thanks, John. That was the type of information I was looking for.
Joe Gray
KA5ZEC
On Mon, Oct 5, 2009 at 11:34 AM, John Ackermann N8UR <jra@febo.com> wrote:
> Here's a reason why you might "want" (as opposed to "need"!) another
> standard beyond the GPSDO.
>
> An Rb or Cs is self-contained. The GPSDO relies on an external factor, the
> presence of GPS. Holdover mode may be good to keep thing close for a while,
> but over the long term the GPSDO without GPS is no better than its crystal
> oscillator.
>
> The Rb is a secondary standard and therefore isn't "correct" by definition,
> but it has a low aging rate and gives you something independent of GPS to
> use for measurements. Monitor it against GPS for a while to learn its
> offset and drift characteristics, and then you can extrapolate its
> performance out over a much longer time than you could with an OCXO.
>
> You may not require that independence, but it gives you additional
> measurement capability. For example, comparing the output of two GPSDO may
> not be meaningful because their frequencies could be correlated by their
> common view of the GPS constellation. Using an Rb reference would eliminate
> that common mode error and reveal information about the GPSDO's short and
> medium term stability that would otherwise be hidden.
>
> John
> ----
> Joseph Gray wrote:
>>
>> I know I don't "need" any of this stuff. I was just wondering what I
>> could do with a rubidium vs what I already have.
>>
>> On Mon, Oct 5, 2009 at 10:58 AM, J. Forster <jfor@quik.com> wrote:
>>>
>>> From an engineering point of view:
>>>
>>> Very few people really "NEED" a Rb or Cs as they are not really doing
>>> anything that requires that level of precision in frequency or accuracy
>>> in
>>> time (me included). These days a BC-221 or WWB is not good enough, but 1
>>> Hz at X-Band is for almost all practical uses.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
JF
J. Forster
Mon, Oct 5, 2009 7:35 PM
A GOOD crystal oscillator will stay w/in parts in 10E10 for days at least.
And I don't mean a DO. Just run it with a continuous comparison to LORAN
or WWVB or GPS and MANUALLY tweek it as needed. A benefit is that you KNOW
when you made an adjustment and not have to worry about random changes
from the oscillator. That's a reason I stopped using mine.
From an engineering point of view, you rarely need to be that close to
frequency in absolute terms. Being very stable is often enough. As with
voltage and current, often what really matters is the ratio, not the
absolute value. Only if you are doing fundamental physics research or
tracking crustal movements is such ultra high precision really needed.
Why do you really care if your clock is off a few nanoseconds, or your
position off a few feet? If you are going to put up a fence, the legally
defined property line is the issue, not what GPS says.
My main gripe w/ Rb and especially Cs is they have very limited lives.
IMO, it's just not worth paying extra money for stability I don't need. If
I ever do need their capabilities, there are several on the shelf, but
until then they can sit unpowered.
FWIW,
-John
==============
Here's a reason why you might "want" (as opposed to "need"!) another
standard beyond the GPSDO.
An Rb or Cs is self-contained. The GPSDO relies on an external factor,
the presence of GPS. Holdover mode may be good to keep thing close for
a while, but over the long term the GPSDO without GPS is no better than
its crystal oscillator.
The Rb is a secondary standard and therefore isn't "correct" by
definition, but it has a low aging rate and gives you something
independent of GPS to use for measurements. Monitor it against GPS for
a while to learn its offset and drift characteristics, and then you can
extrapolate its performance out over a much longer time than you could
with an OCXO.
You may not require that independence, but it gives you additional
measurement capability. For example, comparing the output of two GPSDO
may not be meaningful because their frequencies could be correlated by
their common view of the GPS constellation. Using an Rb reference would
eliminate that common mode error and reveal information about the
GPSDO's short and medium term stability that would otherwise be hidden.
John
A GOOD crystal oscillator will stay w/in parts in 10E10 for days at least.
And I don't mean a DO. Just run it with a continuous comparison to LORAN
or WWVB or GPS and MANUALLY tweek it as needed. A benefit is that you KNOW
when you made an adjustment and not have to worry about random changes
from the oscillator. That's a reason I stopped using mine.
>From an engineering point of view, you rarely need to be that close to
frequency in absolute terms. Being very stable is often enough. As with
voltage and current, often what really matters is the ratio, not the
absolute value. Only if you are doing fundamental physics research or
tracking crustal movements is such ultra high precision really needed.
Why do you really care if your clock is off a few nanoseconds, or your
position off a few feet? If you are going to put up a fence, the legally
defined property line is the issue, not what GPS says.
My main gripe w/ Rb and especially Cs is they have very limited lives.
IMO, it's just not worth paying extra money for stability I don't need. If
I ever do need their capabilities, there are several on the shelf, but
until then they can sit unpowered.
FWIW,
-John
==============
> Here's a reason why you might "want" (as opposed to "need"!) another
> standard beyond the GPSDO.
>
> An Rb or Cs is self-contained. The GPSDO relies on an external factor,
> the presence of GPS. Holdover mode may be good to keep thing close for
> a while, but over the long term the GPSDO without GPS is no better than
> its crystal oscillator.
>
> The Rb is a secondary standard and therefore isn't "correct" by
> definition, but it has a low aging rate and gives you something
> independent of GPS to use for measurements. Monitor it against GPS for
> a while to learn its offset and drift characteristics, and then you can
> extrapolate its performance out over a much longer time than you could
> with an OCXO.
>
> You may not require that independence, but it gives you additional
> measurement capability. For example, comparing the output of two GPSDO
> may not be meaningful because their frequencies could be correlated by
> their common view of the GPS constellation. Using an Rb reference would
> eliminate that common mode error and reveal information about the
> GPSDO's short and medium term stability that would otherwise be hidden.
>
> John
DI
David I. Emery
Mon, Oct 5, 2009 9:46 PM
On Mon, Oct 05, 2009 at 12:35:17PM -0700, J. Forster wrote:
My main gripe w/ Rb and especially Cs is they have very limited lives.
IMO, it's just not worth paying extra money for stability I don't need. If
I ever do need their capabilities, there are several on the shelf, but
until then they can sit unpowered.
The telcom grade (LPRO and similar) Rb's that show up cheap are
supposedly designed for 20 years life continuously operating... and some
have only a few hundred hours on them or less when surplussed. They
should typically last 10-15 years at least in continuous home lab
service and given their price these days the electricity to power them
for that time interval is much more money than the hardware.
Cesiums have REALLY expensive tubes with limited life, but on the
other hand I understand the tubes don't like to be kept turned off... they
deteriorate when sitting on a shelf and may not recover.
--
Dave Emery N1PRE/AE, die@dieconsulting.com DIE Consulting, Weston, Mass 02493
"An empty zombie mind with a forlorn barely readable weatherbeaten
'For Rent' sign still vainly flapping outside on the weed encrusted pole - in
celebration of what could have been, but wasn't and is not to be now either."
On Mon, Oct 05, 2009 at 12:35:17PM -0700, J. Forster wrote:
>
> My main gripe w/ Rb and especially Cs is they have very limited lives.
> IMO, it's just not worth paying extra money for stability I don't need. If
> I ever do need their capabilities, there are several on the shelf, but
> until then they can sit unpowered.
The telcom grade (LPRO and similar) Rb's that show up cheap are
supposedly designed for 20 years life continuously operating... and some
have only a few hundred hours on them or less when surplussed. They
should typically last 10-15 years at least in continuous home lab
service and given their price these days the electricity to power them
for that time interval is much more money than the hardware.
Cesiums have REALLY expensive tubes with limited life, but on the
other hand I understand the tubes don't like to be kept turned off... they
deteriorate when sitting on a shelf and may not recover.
--
Dave Emery N1PRE/AE, die@dieconsulting.com DIE Consulting, Weston, Mass 02493
"An empty zombie mind with a forlorn barely readable weatherbeaten
'For Rent' sign still vainly flapping outside on the weed encrusted pole - in