Temperature accuracy / repeatability / drift of double ovens:
What can be reached here, with a tolerable effort?
Is there anything known about that in the public? mK-numbers?
The application is not a crystal oven this time, but diode lasers
that interrogate Rb atoms and must be kept in lockstep at a certain
beat frequency. The frequency tuning is mostly by temperature;
laser diode current plays a minor role, but probably via changing the
local temperature in the junction and a little bit later in the cavity.
Are there any GoTo thermistors / circuits for stability?
Cheers, Gerhard
(stumbling into new territory)
Temperature accuracy / repeatability / drift of double ovens:
What can be reached here, with a tolerable effort?
Is there anything known about that in the public? mK-numbers?
The application is not a crystal oven this time, but diode lasers
that interrogate Rb atoms and must be kept in lockstep at a certain
beat frequency. The frequency tuning is mostly by temperature;
laser diode current plays a minor role, but probably via changing the
local temperature in the junction and a little bit later in the cavity.
Are there any GoTo thermistors / circuits for stability?
Cheers, Gerhard
(stumbling into new territory)
BG
Bruce Griffiths
Thu, Mar 30, 2023 11:00 PM
Actually the laser diode current also alters the refractive index of the laser diode chip.
Usually, an external cavity is used to stabilise and tune the diode laser operating wavelength in such applications. A cateye ECDL with a tilt tuned narrow bandwidth filter is easier to construct than either Littrow or Littman-Metcalf EDL's using tilt tuned gratings.
If the cavity is reanonably compact then a single oven design similar to that used by the E1938A would perhaps be a better choice.
However laser diode temperature controllers usually employ a TEC to cool the diode rather than a heater.
Whilst some FP laser diodes will operate in single transverse and single longitudinal mode without an external cavity and associated band select filter achieving this together with operation at the desired wavelength can be problematic.
I have a single transverse mode, single longitudinal mode FP laser diode that uses a TEC to control the diode temperature along with a low noise current drive. Both the temperature and operating current are selected so that mode hopping doesnt ocurr in temperature and current island of stability.
Bruce
On 31/03/2023 11:16 NZDT Gerhard Hoffmann via time-nuts time-nuts@lists.febo.com wrote:
Temperature accuracy / repeatability / drift of double ovens:
What can be reached here, with a tolerable effort?
Is there anything known about that in the public? mK-numbers?
The application is not a crystal oven this time, but diode lasers
that interrogate Rb atoms and must be kept in lockstep at a certain
beat frequency. The frequency tuning is mostly by temperature;
laser diode current plays a minor role, but probably via changing the
local temperature in the junction and a little bit later in the cavity.
Are there any GoTo thermistors / circuits for stability?
Cheers, Gerhard
(stumbling into new territory)
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Actually the laser diode current also alters the refractive index of the laser diode chip.
Usually, an external cavity is used to stabilise and tune the diode laser operating wavelength in such applications. A cateye ECDL with a tilt tuned narrow bandwidth filter is easier to construct than either Littrow or Littman-Metcalf EDL's using tilt tuned gratings.
If the cavity is reanonably compact then a single oven design similar to that used by the E1938A would perhaps be a better choice.
However laser diode temperature controllers usually employ a TEC to cool the diode rather than a heater.
Whilst some FP laser diodes will operate in single transverse and single longitudinal mode without an external cavity and associated band select filter achieving this together with operation at the desired wavelength can be problematic.
I have a single transverse mode, single longitudinal mode FP laser diode that uses a TEC to control the diode temperature along with a low noise current drive. Both the temperature and operating current are selected so that mode hopping doesnt ocurr in temperature and current island of stability.
Bruce
> On 31/03/2023 11:16 NZDT Gerhard Hoffmann via time-nuts <time-nuts@lists.febo.com> wrote:
>
>
> Temperature accuracy / repeatability / drift of double ovens:
>
> What can be reached here, with a tolerable effort?
> Is there anything known about that in the public? mK-numbers?
>
> The application is not a crystal oven this time, but diode lasers
> that interrogate Rb atoms and must be kept in lockstep at a certain
> beat frequency. The frequency tuning is mostly by temperature;
> laser diode current plays a minor role, but probably via changing the
> local temperature in the junction and a little bit later in the cavity.
>
> Are there any GoTo thermistors / circuits for stability?
>
> Cheers, Gerhard
> (stumbling into new territory)
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
BC
Bob Camp
Thu, Mar 30, 2023 11:01 PM
Hi
Simple answers (and likely not much help):
Thermal gain on a single oven is likely in the 300 to 600 range.
Double oven boosts that by 10 to 30X
Thermal gain in this case: take the ambient change divide it by the thermal gain
and you get the oven temperature change.
Long term stability:
low enough that it does not impact aging. ( = that’s how you test it). Since that gets back
to the specific thermistor being used, you are into the ‘likely not much help” range.
Best guess info: The crystal + circuit is < 1x10^-9 / C in the vicinity of the turn. The operating
point is offset from turn to minimize the combination of crystal + circuit. Just what the net
is ….
Still, if you can “see” aging at the parts in the (low?) 10^-11 range, it sort of kind of gets
you to a number. Maybe 0.01C / day.
Since you are canceling the circuit with the crystal, if things got to far off, a year later the
temperature performance would not be any good. That likely gets you to something
in the < 1C per year range.
Bob
On Mar 30, 2023, at 6:16 PM, Gerhard Hoffmann via time-nuts time-nuts@lists.febo.com wrote:
Temperature accuracy / repeatability / drift of double ovens:
What can be reached here, with a tolerable effort?
Is there anything known about that in the public? mK-numbers?
The application is not a crystal oven this time, but diode lasers
that interrogate Rb atoms and must be kept in lockstep at a certain
beat frequency. The frequency tuning is mostly by temperature;
laser diode current plays a minor role, but probably via changing the
local temperature in the junction and a little bit later in the cavity.
Are there any GoTo thermistors / circuits for stability?
Cheers, Gerhard
(stumbling into new territory)
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Hi
Simple answers (and likely not much help):
Thermal gain on a single oven is likely in the 300 to 600 range.
Double oven boosts that by 10 to 30X
Thermal gain in this case: take the ambient change divide it by the thermal gain
and you get the oven temperature change.
Long term stability:
low enough that it does not impact aging. ( = that’s how you test it). Since that gets back
to the specific thermistor being used, you are into the ‘likely not much help” range.
Best guess info: The crystal + circuit is < 1x10^-9 / C in the vicinity of the turn. The operating
point is offset from turn to minimize the combination of crystal + circuit. Just what the net
is ….
Still, if you can “see” aging at the parts in the (low?) 10^-11 range, it sort of kind of gets
you to a number. Maybe 0.01C / day.
Since you are canceling the circuit with the crystal, if things got to far off, a year later the
temperature performance would not be any good. That likely gets you to something
in the < 1C per year range.
Bob
> On Mar 30, 2023, at 6:16 PM, Gerhard Hoffmann via time-nuts <time-nuts@lists.febo.com> wrote:
>
> Temperature accuracy / repeatability / drift of double ovens:
>
> What can be reached here, with a tolerable effort?
> Is there anything known about that in the public? mK-numbers?
>
> The application is not a crystal oven this time, but diode lasers
> that interrogate Rb atoms and must be kept in lockstep at a certain
> beat frequency. The frequency tuning is mostly by temperature;
> laser diode current plays a minor role, but probably via changing the
> local temperature in the junction and a little bit later in the cavity.
>
> Are there any GoTo thermistors / circuits for stability?
>
> Cheers, Gerhard
> (stumbling into new territory)
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
MD
Magnus Danielson
Thu, Mar 30, 2023 11:57 PM
The DFB lasers you use for this typically has a thermistor and
termoelectical cooler to form a control loop with external components,
just as for any other DWDM DFB laser. Tune the setpoint of this to lock
the laser to the rubidium. The laser becomes warm, so you stabilize it
by how much you cool it.
Further, it has a photodiode to set a loop to stabilize the current and
thus the intensity of the laser, which often is of interest to stabilize
for clock transitions.
Add modulation to the laser for sidebands if that is relevant.
One DFB datasheet I found had typical 0.06 nm / K in temperature
sensitivity and typ 0.003 nm / mA in current sensitivity. Thus, it could
be wise to "leak" the current-steering to the temperature to compensate
for current effect in feed forward rather than indirect lock.
Look at the available DFB lasermodules, I think it should become fairly
clear quite quickly as you look at datasheets etc. Exactly which laser
to use depends on power and bandwidth that fit a particular use, so it
boils down to what you end up needing it to do.
Cheers,
Magnus
On 2023-03-31 00:16, Gerhard Hoffmann via time-nuts wrote:
Temperature accuracy / repeatability / drift of double ovens:
What can be reached here, with a tolerable effort?
Is there anything known about that in the public? mK-numbers?
The application is not a crystal oven this time, but diode lasers
that interrogate Rb atoms and must be kept in lockstep at a certain
beat frequency. The frequency tuning is mostly by temperature;
laser diode current plays a minor role, but probably via changing the
local temperature in the junction and a little bit later in the cavity.
Are there any GoTo thermistors / circuits for stability?
Cheers, Gerhard
(stumbling into new territory)
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
The DFB lasers you use for this typically has a thermistor and
termoelectical cooler to form a control loop with external components,
just as for any other DWDM DFB laser. Tune the setpoint of this to lock
the laser to the rubidium. The laser becomes warm, so you stabilize it
by how much you cool it.
Further, it has a photodiode to set a loop to stabilize the current and
thus the intensity of the laser, which often is of interest to stabilize
for clock transitions.
Add modulation to the laser for sidebands if that is relevant.
One DFB datasheet I found had typical 0.06 nm / K in temperature
sensitivity and typ 0.003 nm / mA in current sensitivity. Thus, it could
be wise to "leak" the current-steering to the temperature to compensate
for current effect in feed forward rather than indirect lock.
Look at the available DFB lasermodules, I think it should become fairly
clear quite quickly as you look at datasheets etc. Exactly which laser
to use depends on power and bandwidth that fit a particular use, so it
boils down to what you end up needing it to do.
Cheers,
Magnus
On 2023-03-31 00:16, Gerhard Hoffmann via time-nuts wrote:
> Temperature accuracy / repeatability / drift of double ovens:
>
> What can be reached here, with a tolerable effort?
> Is there anything known about that in the public? mK-numbers?
>
> The application is not a crystal oven this time, but diode lasers
> that interrogate Rb atoms and must be kept in lockstep at a certain
> beat frequency. The frequency tuning is mostly by temperature;
> laser diode current plays a minor role, but probably via changing the
> local temperature in the junction and a little bit later in the cavity.
>
> Are there any GoTo thermistors / circuits for stability?
>
> Cheers, Gerhard
> (stumbling into new territory)
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
EM
Ed Marciniak
Fri, Mar 31, 2023 6:13 PM
Linear technology made a thermoelectric full H bridge temperature controller. At the time, I’d like to say it came as a lot of two components. The second was an op-amp that may have been selected or trimmed unless there was a good reason to include it with samples (maybe less commonly available).
It was supposed to be able to achieve 10 millikelvin stability easily, and with more care in design, 1 millikelvin stability was supposed to be not too difficult to achieve.
It was a moderately expensive part…about $20 for the two components. A handful of passives and a a pair of N-channel and a pair of P-channel MOSFETs were required.
If memory serves correctly it was also switched mode rather than linear.
Get Outlook for iOShttps://aka.ms/o0ukef
From: Bob Camp via time-nuts time-nuts@lists.febo.com
Sent: Thursday, March 30, 2023 6:01:55 PM
To: Discussion of precise time and frequency measurement time-nuts@lists.febo.com
Cc: Bob Camp kb8tq@n1k.org
Subject: [time-nuts] Re: Temperature accuracy / repeatability / drift of double ovens
Hi
Simple answers (and likely not much help):
Thermal gain on a single oven is likely in the 300 to 600 range.
Double oven boosts that by 10 to 30X
Thermal gain in this case: take the ambient change divide it by the thermal gain
and you get the oven temperature change.
Long term stability:
low enough that it does not impact aging. ( = that’s how you test it). Since that gets back
to the specific thermistor being used, you are into the ‘likely not much help” range.
Best guess info: The crystal + circuit is < 1x10^-9 / C in the vicinity of the turn. The operating
point is offset from turn to minimize the combination of crystal + circuit. Just what the net
is ….
Still, if you can “see” aging at the parts in the (low?) 10^-11 range, it sort of kind of gets
you to a number. Maybe 0.01C / day.
Since you are canceling the circuit with the crystal, if things got to far off, a year later the
temperature performance would not be any good. That likely gets you to something
in the < 1C per year range.
Bob
On Mar 30, 2023, at 6:16 PM, Gerhard Hoffmann via time-nuts time-nuts@lists.febo.com wrote:
Temperature accuracy / repeatability / drift of double ovens:
What can be reached here, with a tolerable effort?
Is there anything known about that in the public? mK-numbers?
The application is not a crystal oven this time, but diode lasers
that interrogate Rb atoms and must be kept in lockstep at a certain
beat frequency. The frequency tuning is mostly by temperature;
laser diode current plays a minor role, but probably via changing the
local temperature in the junction and a little bit later in the cavity.
Are there any GoTo thermistors / circuits for stability?
Cheers, Gerhard
(stumbling into new territory)
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Linear technology made a thermoelectric full H bridge temperature controller. At the time, I’d like to say it came as a lot of two components. The second was an op-amp that may have been selected or trimmed unless there was a good reason to include it with samples (maybe less commonly available).
It was supposed to be able to achieve 10 millikelvin stability easily, and with more care in design, 1 millikelvin stability was supposed to be not too difficult to achieve.
It was a moderately expensive part…about $20 for the two components. A handful of passives and a a pair of N-channel and a pair of P-channel MOSFETs were required.
If memory serves correctly it was also switched mode rather than linear.
Get Outlook for iOS<https://aka.ms/o0ukef>
________________________________
From: Bob Camp via time-nuts <time-nuts@lists.febo.com>
Sent: Thursday, March 30, 2023 6:01:55 PM
To: Discussion of precise time and frequency measurement <time-nuts@lists.febo.com>
Cc: Bob Camp <kb8tq@n1k.org>
Subject: [time-nuts] Re: Temperature accuracy / repeatability / drift of double ovens
Hi
Simple answers (and likely not much help):
Thermal gain on a single oven is likely in the 300 to 600 range.
Double oven boosts that by 10 to 30X
Thermal gain in this case: take the ambient change divide it by the thermal gain
and you get the oven temperature change.
Long term stability:
low enough that it does not impact aging. ( = that’s how you test it). Since that gets back
to the specific thermistor being used, you are into the ‘likely not much help” range.
Best guess info: The crystal + circuit is < 1x10^-9 / C in the vicinity of the turn. The operating
point is offset from turn to minimize the combination of crystal + circuit. Just what the net
is ….
Still, if you can “see” aging at the parts in the (low?) 10^-11 range, it sort of kind of gets
you to a number. Maybe 0.01C / day.
Since you are canceling the circuit with the crystal, if things got to far off, a year later the
temperature performance would not be any good. That likely gets you to something
in the < 1C per year range.
Bob
> On Mar 30, 2023, at 6:16 PM, Gerhard Hoffmann via time-nuts <time-nuts@lists.febo.com> wrote:
>
> Temperature accuracy / repeatability / drift of double ovens:
>
> What can be reached here, with a tolerable effort?
> Is there anything known about that in the public? mK-numbers?
>
> The application is not a crystal oven this time, but diode lasers
> that interrogate Rb atoms and must be kept in lockstep at a certain
> beat frequency. The frequency tuning is mostly by temperature;
> laser diode current plays a minor role, but probably via changing the
> local temperature in the junction and a little bit later in the cavity.
>
> Are there any GoTo thermistors / circuits for stability?
>
> Cheers, Gerhard
> (stumbling into new territory)
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
BC
Bob Camp
Fri, Mar 31, 2023 7:03 PM
Hi
There are a lot of ways to rate this sort of thing. One is to look at the level of control
at the temperature sensor. It might be a thermistor, it could easily be something else.
For a long time OCXO designers tossed around this as their “control level”.
You would look at the variation on the thermistor bridge over some period. Convert
the volts to resistance and then to C, F, or K. Folks would get numbers like “12 micro
kelvin” as a result.
A fairly simple step back, scratch head, and think about it sort of analysis ultimately
showed this to be a bit useless or any sort of real performance evaluation.
What matters is the level of temperature control over some volume ( the inside of
an OCXO or ….). Real world devices are not completely located at a single point.
It’s a bit amazing just how much things change over a CM or so of distance in an
enclosure. Toss in variable heat sources and it really gets wild.
One route down this rabbit hole:
http://www.karlquist.com/oven.pdf
One also might note that the author is floating around here on the list.
Bob
On Mar 31, 2023, at 2:13 PM, Ed Marciniak ed@nb0m.org wrote:
Linear technology made a thermoelectric full H bridge temperature controller. At the time, I’d like to say it came as a lot of two components. The second was an op-amp that may have been selected or trimmed unless there was a good reason to include it with samples (maybe less commonly available).
It was supposed to be able to achieve 10 millikelvin stability easily, and with more care in design, 1 millikelvin stability was supposed to be not too difficult to achieve.
It was a moderately expensive part…about $20 for the two components. A handful of passives and a a pair of N-channel and a pair of P-channel MOSFETs were required.
If memory serves correctly it was also switched mode rather than linear.
Get Outlook for iOS https://aka.ms/o0ukef
From: Bob Camp via time-nuts time-nuts@lists.febo.com
Sent: Thursday, March 30, 2023 6:01:55 PM
To: Discussion of precise time and frequency measurement time-nuts@lists.febo.com
Cc: Bob Camp kb8tq@n1k.org
Subject: [time-nuts] Re: Temperature accuracy / repeatability / drift of double ovens
Hi
Simple answers (and likely not much help):
Thermal gain on a single oven is likely in the 300 to 600 range.
Double oven boosts that by 10 to 30X
Thermal gain in this case: take the ambient change divide it by the thermal gain
and you get the oven temperature change.
Long term stability:
low enough that it does not impact aging. ( = that’s how you test it). Since that gets back
to the specific thermistor being used, you are into the ‘likely not much help” range.
Best guess info: The crystal + circuit is < 1x10^-9 / C in the vicinity of the turn. The operating
point is offset from turn to minimize the combination of crystal + circuit. Just what the net
is ….
Still, if you can “see” aging at the parts in the (low?) 10^-11 range, it sort of kind of gets
you to a number. Maybe 0.01C / day.
Since you are canceling the circuit with the crystal, if things got to far off, a year later the
temperature performance would not be any good. That likely gets you to something
in the < 1C per year range.
Bob
On Mar 30, 2023, at 6:16 PM, Gerhard Hoffmann via time-nuts time-nuts@lists.febo.com wrote:
Temperature accuracy / repeatability / drift of double ovens:
What can be reached here, with a tolerable effort?
Is there anything known about that in the public? mK-numbers?
The application is not a crystal oven this time, but diode lasers
that interrogate Rb atoms and must be kept in lockstep at a certain
beat frequency. The frequency tuning is mostly by temperature;
laser diode current plays a minor role, but probably via changing the
local temperature in the junction and a little bit later in the cavity.
Are there any GoTo thermistors / circuits for stability?
Cheers, Gerhard
(stumbling into new territory)
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Hi
There are a lot of ways to rate this sort of thing. One is to look at the level of control
at the temperature sensor. It might be a thermistor, it could easily be something else.
For a long time OCXO designers tossed around this as their “control level”.
You would look at the variation on the thermistor bridge over some period. Convert
the volts to resistance and then to C, F, or K. Folks would get numbers like “12 micro
kelvin” as a result.
A fairly simple step back, scratch head, and think about it sort of analysis ultimately
showed this to be a bit useless or any sort of real performance evaluation.
What matters is the level of temperature control over some volume ( the inside of
an OCXO or ….). Real world devices are not completely located at a single point.
It’s a bit amazing just how much things change over a CM or so of distance in an
enclosure. Toss in variable heat sources and it really gets wild.
One route down this rabbit hole:
http://www.karlquist.com/oven.pdf
One also might note that the author is floating around here on the list.
Bob
> On Mar 31, 2023, at 2:13 PM, Ed Marciniak <ed@nb0m.org> wrote:
>
> Linear technology made a thermoelectric full H bridge temperature controller. At the time, I’d like to say it came as a lot of two components. The second was an op-amp that may have been selected or trimmed unless there was a good reason to include it with samples (maybe less commonly available).
>
> It was supposed to be able to achieve 10 millikelvin stability easily, and with more care in design, 1 millikelvin stability was supposed to be not too difficult to achieve.
>
> It was a moderately expensive part…about $20 for the two components. A handful of passives and a a pair of N-channel and a pair of P-channel MOSFETs were required.
>
> If memory serves correctly it was also switched mode rather than linear.
>
> Get Outlook for iOS <https://aka.ms/o0ukef>
> From: Bob Camp via time-nuts <time-nuts@lists.febo.com>
> Sent: Thursday, March 30, 2023 6:01:55 PM
> To: Discussion of precise time and frequency measurement <time-nuts@lists.febo.com>
> Cc: Bob Camp <kb8tq@n1k.org>
> Subject: [time-nuts] Re: Temperature accuracy / repeatability / drift of double ovens
>
> Hi
>
> Simple answers (and likely not much help):
>
> Thermal gain on a single oven is likely in the 300 to 600 range.
>
> Double oven boosts that by 10 to 30X
>
> Thermal gain in this case: take the ambient change divide it by the thermal gain
> and you get the oven temperature change.
>
> Long term stability:
>
> low enough that it does not impact aging. ( = that’s how you test it). Since that gets back
> to the specific thermistor being used, you are into the ‘likely not much help” range.
>
> Best guess info: The crystal + circuit is < 1x10^-9 / C in the vicinity of the turn. The operating
> point is offset from turn to minimize the combination of crystal + circuit. Just what the net
> is ….
>
> Still, if you can “see” aging at the parts in the (low?) 10^-11 range, it sort of kind of gets
> you to a number. Maybe 0.01C / day.
>
> Since you are canceling the circuit with the crystal, if things got to far off, a year later the
> temperature performance would not be any good. That likely gets you to something
> in the < 1C per year range.
>
> Bob
>
> > On Mar 30, 2023, at 6:16 PM, Gerhard Hoffmann via time-nuts <time-nuts@lists.febo.com> wrote:
> >
> > Temperature accuracy / repeatability / drift of double ovens:
> >
> > What can be reached here, with a tolerable effort?
> > Is there anything known about that in the public? mK-numbers?
> >
> > The application is not a crystal oven this time, but diode lasers
> > that interrogate Rb atoms and must be kept in lockstep at a certain
> > beat frequency. The frequency tuning is mostly by temperature;
> > laser diode current plays a minor role, but probably via changing the
> > local temperature in the junction and a little bit later in the cavity.
> >
> > Are there any GoTo thermistors / circuits for stability?
> >
> > Cheers, Gerhard
> > (stumbling into new territory)
> > _______________________________________________
> > time-nuts mailing list -- time-nuts@lists.febo.com
> > To unsubscribe send an email to time-nuts-leave@lists.febo.com
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
MD
Magnus Danielson
Fri, Mar 31, 2023 7:26 PM
Hi,
There is a few things which differs. The oven for crystals is quite
different from the cooling of DFB lasers. You regulate your built in
Peltier element. Also, the temperature gradient over the crystal blank
isn't the same as the temperature of the DFB laser diode.
So, there is a bit difference in these rabbit holes.
Cheers,
Magnus
On 2023-03-31 21:03, Bob Camp via time-nuts wrote:
Hi
There are a lot of ways to rate this sort of thing. One is to look at the level of control
at the temperature sensor. It might be a thermistor, it could easily be something else.
For a long time OCXO designers tossed around this as their “control level”.
You would look at the variation on the thermistor bridge over some period. Convert
the volts to resistance and then to C, F, or K. Folks would get numbers like “12 micro
kelvin” as a result.
A fairly simple step back, scratch head, and think about it sort of analysis ultimately
showed this to be a bit useless or any sort of real performance evaluation.
What matters is the level of temperature control over some volume ( the inside of
an OCXO or ….). Real world devices are not completely located at a single point.
It’s a bit amazing just how much things change over a CM or so of distance in an
enclosure. Toss in variable heat sources and it really gets wild.
One route down this rabbit hole:
http://www.karlquist.com/oven.pdf
One also might note that the author is floating around here on the list.
Bob
On Mar 31, 2023, at 2:13 PM, Ed Marciniak ed@nb0m.org wrote:
Linear technology made a thermoelectric full H bridge temperature controller. At the time, I’d like to say it came as a lot of two components. The second was an op-amp that may have been selected or trimmed unless there was a good reason to include it with samples (maybe less commonly available).
It was supposed to be able to achieve 10 millikelvin stability easily, and with more care in design, 1 millikelvin stability was supposed to be not too difficult to achieve.
It was a moderately expensive part…about $20 for the two components. A handful of passives and a a pair of N-channel and a pair of P-channel MOSFETs were required.
If memory serves correctly it was also switched mode rather than linear.
Get Outlook for iOS https://aka.ms/o0ukef
From: Bob Camp via time-nuts time-nuts@lists.febo.com
Sent: Thursday, March 30, 2023 6:01:55 PM
To: Discussion of precise time and frequency measurement time-nuts@lists.febo.com
Cc: Bob Camp kb8tq@n1k.org
Subject: [time-nuts] Re: Temperature accuracy / repeatability / drift of double ovens
Hi
Simple answers (and likely not much help):
Thermal gain on a single oven is likely in the 300 to 600 range.
Double oven boosts that by 10 to 30X
Thermal gain in this case: take the ambient change divide it by the thermal gain
and you get the oven temperature change.
Long term stability:
low enough that it does not impact aging. ( = that’s how you test it). Since that gets back
to the specific thermistor being used, you are into the ‘likely not much help” range.
Best guess info: The crystal + circuit is < 1x10^-9 / C in the vicinity of the turn. The operating
point is offset from turn to minimize the combination of crystal + circuit. Just what the net
is ….
Still, if you can “see” aging at the parts in the (low?) 10^-11 range, it sort of kind of gets
you to a number. Maybe 0.01C / day.
Since you are canceling the circuit with the crystal, if things got to far off, a year later the
temperature performance would not be any good. That likely gets you to something
in the < 1C per year range.
Bob
On Mar 30, 2023, at 6:16 PM, Gerhard Hoffmann via time-nuts time-nuts@lists.febo.com wrote:
Temperature accuracy / repeatability / drift of double ovens:
What can be reached here, with a tolerable effort?
Is there anything known about that in the public? mK-numbers?
The application is not a crystal oven this time, but diode lasers
that interrogate Rb atoms and must be kept in lockstep at a certain
beat frequency. The frequency tuning is mostly by temperature;
laser diode current plays a minor role, but probably via changing the
local temperature in the junction and a little bit later in the cavity.
Are there any GoTo thermistors / circuits for stability?
Cheers, Gerhard
(stumbling into new territory)
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Hi,
There is a few things which differs. The oven for crystals is quite
different from the cooling of DFB lasers. You regulate your built in
Peltier element. Also, the temperature gradient over the crystal blank
isn't the same as the temperature of the DFB laser diode.
So, there is a bit difference in these rabbit holes.
Cheers,
Magnus
On 2023-03-31 21:03, Bob Camp via time-nuts wrote:
> Hi
>
> There are a lot of ways to rate this sort of thing. One is to look at the level of control
> at the temperature sensor. It might be a thermistor, it could easily be something else.
> For a long time OCXO designers tossed around this as their “control level”.
>
> You would look at the variation on the thermistor bridge over some period. Convert
> the volts to resistance and then to C, F, or K. Folks would get numbers like “12 micro
> kelvin” as a result.
>
> A fairly simple step back, scratch head, and think about it sort of analysis ultimately
> showed this to be a bit useless or any sort of real performance evaluation.
>
> What matters is the level of temperature control over some volume ( the inside of
> an OCXO or ….). Real world devices are not completely located at a single point.
> It’s a bit amazing just how much things change over a CM or so of distance in an
> enclosure. Toss in variable heat sources and it really gets wild.
>
> One route down this rabbit hole:
>
> http://www.karlquist.com/oven.pdf
>
> One also might note that the author is floating around here on the list.
>
> Bob
>
>> On Mar 31, 2023, at 2:13 PM, Ed Marciniak <ed@nb0m.org> wrote:
>>
>> Linear technology made a thermoelectric full H bridge temperature controller. At the time, I’d like to say it came as a lot of two components. The second was an op-amp that may have been selected or trimmed unless there was a good reason to include it with samples (maybe less commonly available).
>>
>> It was supposed to be able to achieve 10 millikelvin stability easily, and with more care in design, 1 millikelvin stability was supposed to be not too difficult to achieve.
>>
>> It was a moderately expensive part…about $20 for the two components. A handful of passives and a a pair of N-channel and a pair of P-channel MOSFETs were required.
>>
>> If memory serves correctly it was also switched mode rather than linear.
>>
>> Get Outlook for iOS <https://aka.ms/o0ukef>
>> From: Bob Camp via time-nuts <time-nuts@lists.febo.com>
>> Sent: Thursday, March 30, 2023 6:01:55 PM
>> To: Discussion of precise time and frequency measurement <time-nuts@lists.febo.com>
>> Cc: Bob Camp <kb8tq@n1k.org>
>> Subject: [time-nuts] Re: Temperature accuracy / repeatability / drift of double ovens
>>
>> Hi
>>
>> Simple answers (and likely not much help):
>>
>> Thermal gain on a single oven is likely in the 300 to 600 range.
>>
>> Double oven boosts that by 10 to 30X
>>
>> Thermal gain in this case: take the ambient change divide it by the thermal gain
>> and you get the oven temperature change.
>>
>> Long term stability:
>>
>> low enough that it does not impact aging. ( = that’s how you test it). Since that gets back
>> to the specific thermistor being used, you are into the ‘likely not much help” range.
>>
>> Best guess info: The crystal + circuit is < 1x10^-9 / C in the vicinity of the turn. The operating
>> point is offset from turn to minimize the combination of crystal + circuit. Just what the net
>> is ….
>>
>> Still, if you can “see” aging at the parts in the (low?) 10^-11 range, it sort of kind of gets
>> you to a number. Maybe 0.01C / day.
>>
>> Since you are canceling the circuit with the crystal, if things got to far off, a year later the
>> temperature performance would not be any good. That likely gets you to something
>> in the < 1C per year range.
>>
>> Bob
>>
>>> On Mar 30, 2023, at 6:16 PM, Gerhard Hoffmann via time-nuts <time-nuts@lists.febo.com> wrote:
>>>
>>> Temperature accuracy / repeatability / drift of double ovens:
>>>
>>> What can be reached here, with a tolerable effort?
>>> Is there anything known about that in the public? mK-numbers?
>>>
>>> The application is not a crystal oven this time, but diode lasers
>>> that interrogate Rb atoms and must be kept in lockstep at a certain
>>> beat frequency. The frequency tuning is mostly by temperature;
>>> laser diode current plays a minor role, but probably via changing the
>>> local temperature in the junction and a little bit later in the cavity.
>>>
>>> Are there any GoTo thermistors / circuits for stability?
>>>
>>> Cheers, Gerhard
>>> (stumbling into new territory)
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts@lists.febo.com
>>> To unsubscribe send an email to time-nuts-leave@lists.febo.com
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@lists.febo.com
>> To unsubscribe send an email to time-nuts-leave@lists.febo.com
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
BC
Bob Camp
Fri, Mar 31, 2023 9:17 PM
Hi
Since we started the thread with a question about double ovens …. I’m guessing that the
structure being considered is some sort of oven in a oven sort of thing. In that case, you
are right into a very similar rabbit hole as the OCXO.
Bob
On Mar 31, 2023, at 3:26 PM, Magnus Danielson via time-nuts time-nuts@lists.febo.com wrote:
Hi,
There is a few things which differs. The oven for crystals is quite different from the cooling of DFB lasers. You regulate your built in Peltier element. Also, the temperature gradient over the crystal blank isn't the same as the temperature of the DFB laser diode.
So, there is a bit difference in these rabbit holes.
Cheers,
Magnus
On 2023-03-31 21:03, Bob Camp via time-nuts wrote:
Hi
There are a lot of ways to rate this sort of thing. One is to look at the level of control
at the temperature sensor. It might be a thermistor, it could easily be something else.
For a long time OCXO designers tossed around this as their “control level”.
You would look at the variation on the thermistor bridge over some period. Convert
the volts to resistance and then to C, F, or K. Folks would get numbers like “12 micro
kelvin” as a result.
A fairly simple step back, scratch head, and think about it sort of analysis ultimately
showed this to be a bit useless or any sort of real performance evaluation.
What matters is the level of temperature control over some volume ( the inside of
an OCXO or ….). Real world devices are not completely located at a single point.
It’s a bit amazing just how much things change over a CM or so of distance in an
enclosure. Toss in variable heat sources and it really gets wild.
One route down this rabbit hole:
http://www.karlquist.com/oven.pdf
One also might note that the author is floating around here on the list.
Bob
On Mar 31, 2023, at 2:13 PM, Ed Marciniak ed@nb0m.org wrote:
Linear technology made a thermoelectric full H bridge temperature controller. At the time, I’d like to say it came as a lot of two components. The second was an op-amp that may have been selected or trimmed unless there was a good reason to include it with samples (maybe less commonly available).
It was supposed to be able to achieve 10 millikelvin stability easily, and with more care in design, 1 millikelvin stability was supposed to be not too difficult to achieve.
It was a moderately expensive part…about $20 for the two components. A handful of passives and a a pair of N-channel and a pair of P-channel MOSFETs were required.
If memory serves correctly it was also switched mode rather than linear.
Get Outlook for iOS https://aka.ms/o0ukef
From: Bob Camp via time-nuts time-nuts@lists.febo.com
Sent: Thursday, March 30, 2023 6:01:55 PM
To: Discussion of precise time and frequency measurement time-nuts@lists.febo.com
Cc: Bob Camp kb8tq@n1k.org
Subject: [time-nuts] Re: Temperature accuracy / repeatability / drift of double ovens
Hi
Simple answers (and likely not much help):
Thermal gain on a single oven is likely in the 300 to 600 range.
Double oven boosts that by 10 to 30X
Thermal gain in this case: take the ambient change divide it by the thermal gain
and you get the oven temperature change.
Long term stability:
low enough that it does not impact aging. ( = that’s how you test it). Since that gets back
to the specific thermistor being used, you are into the ‘likely not much help” range.
Best guess info: The crystal + circuit is < 1x10^-9 / C in the vicinity of the turn. The operating
point is offset from turn to minimize the combination of crystal + circuit. Just what the net
is ….
Still, if you can “see” aging at the parts in the (low?) 10^-11 range, it sort of kind of gets
you to a number. Maybe 0.01C / day.
Since you are canceling the circuit with the crystal, if things got to far off, a year later the
temperature performance would not be any good. That likely gets you to something
in the < 1C per year range.
Bob
On Mar 30, 2023, at 6:16 PM, Gerhard Hoffmann via time-nuts time-nuts@lists.febo.com wrote:
Temperature accuracy / repeatability / drift of double ovens:
What can be reached here, with a tolerable effort?
Is there anything known about that in the public? mK-numbers?
The application is not a crystal oven this time, but diode lasers
that interrogate Rb atoms and must be kept in lockstep at a certain
beat frequency. The frequency tuning is mostly by temperature;
laser diode current plays a minor role, but probably via changing the
local temperature in the junction and a little bit later in the cavity.
Are there any GoTo thermistors / circuits for stability?
Cheers, Gerhard
(stumbling into new territory)
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Hi
Since we started the thread with a question about double ovens …. I’m guessing that the
structure being considered is some sort of oven in a oven sort of thing. In that case, you
are right into a very similar rabbit hole as the OCXO.
Bob
> On Mar 31, 2023, at 3:26 PM, Magnus Danielson via time-nuts <time-nuts@lists.febo.com> wrote:
>
> Hi,
>
> There is a few things which differs. The oven for crystals is quite different from the cooling of DFB lasers. You regulate your built in Peltier element. Also, the temperature gradient over the crystal blank isn't the same as the temperature of the DFB laser diode.
>
> So, there is a bit difference in these rabbit holes.
>
> Cheers,
> Magnus
>
> On 2023-03-31 21:03, Bob Camp via time-nuts wrote:
>> Hi
>>
>> There are a lot of ways to rate this sort of thing. One is to look at the level of control
>> at the temperature sensor. It might be a thermistor, it could easily be something else.
>> For a long time OCXO designers tossed around this as their “control level”.
>>
>> You would look at the variation on the thermistor bridge over some period. Convert
>> the volts to resistance and then to C, F, or K. Folks would get numbers like “12 micro
>> kelvin” as a result.
>>
>> A fairly simple step back, scratch head, and think about it sort of analysis ultimately
>> showed this to be a bit useless or any sort of real performance evaluation.
>>
>> What matters is the level of temperature control over some volume ( the inside of
>> an OCXO or ….). Real world devices are not completely located at a single point.
>> It’s a bit amazing just how much things change over a CM or so of distance in an
>> enclosure. Toss in variable heat sources and it really gets wild.
>>
>> One route down this rabbit hole:
>>
>> http://www.karlquist.com/oven.pdf
>>
>> One also might note that the author is floating around here on the list.
>>
>> Bob
>>
>>> On Mar 31, 2023, at 2:13 PM, Ed Marciniak <ed@nb0m.org> wrote:
>>>
>>> Linear technology made a thermoelectric full H bridge temperature controller. At the time, I’d like to say it came as a lot of two components. The second was an op-amp that may have been selected or trimmed unless there was a good reason to include it with samples (maybe less commonly available).
>>>
>>> It was supposed to be able to achieve 10 millikelvin stability easily, and with more care in design, 1 millikelvin stability was supposed to be not too difficult to achieve.
>>>
>>> It was a moderately expensive part…about $20 for the two components. A handful of passives and a a pair of N-channel and a pair of P-channel MOSFETs were required.
>>>
>>> If memory serves correctly it was also switched mode rather than linear.
>>>
>>> Get Outlook for iOS <https://aka.ms/o0ukef>
>>> From: Bob Camp via time-nuts <time-nuts@lists.febo.com>
>>> Sent: Thursday, March 30, 2023 6:01:55 PM
>>> To: Discussion of precise time and frequency measurement <time-nuts@lists.febo.com>
>>> Cc: Bob Camp <kb8tq@n1k.org>
>>> Subject: [time-nuts] Re: Temperature accuracy / repeatability / drift of double ovens
>>> Hi
>>>
>>> Simple answers (and likely not much help):
>>>
>>> Thermal gain on a single oven is likely in the 300 to 600 range.
>>>
>>> Double oven boosts that by 10 to 30X
>>>
>>> Thermal gain in this case: take the ambient change divide it by the thermal gain
>>> and you get the oven temperature change.
>>>
>>> Long term stability:
>>>
>>> low enough that it does not impact aging. ( = that’s how you test it). Since that gets back
>>> to the specific thermistor being used, you are into the ‘likely not much help” range.
>>>
>>> Best guess info: The crystal + circuit is < 1x10^-9 / C in the vicinity of the turn. The operating
>>> point is offset from turn to minimize the combination of crystal + circuit. Just what the net
>>> is ….
>>>
>>> Still, if you can “see” aging at the parts in the (low?) 10^-11 range, it sort of kind of gets
>>> you to a number. Maybe 0.01C / day.
>>>
>>> Since you are canceling the circuit with the crystal, if things got to far off, a year later the
>>> temperature performance would not be any good. That likely gets you to something
>>> in the < 1C per year range.
>>>
>>> Bob
>>>
>>>> On Mar 30, 2023, at 6:16 PM, Gerhard Hoffmann via time-nuts <time-nuts@lists.febo.com> wrote:
>>>>
>>>> Temperature accuracy / repeatability / drift of double ovens:
>>>>
>>>> What can be reached here, with a tolerable effort?
>>>> Is there anything known about that in the public? mK-numbers?
>>>>
>>>> The application is not a crystal oven this time, but diode lasers
>>>> that interrogate Rb atoms and must be kept in lockstep at a certain
>>>> beat frequency. The frequency tuning is mostly by temperature;
>>>> laser diode current plays a minor role, but probably via changing the
>>>> local temperature in the junction and a little bit later in the cavity.
>>>>
>>>> Are there any GoTo thermistors / circuits for stability?
>>>>
>>>> Cheers, Gerhard
>>>> (stumbling into new territory)
>>>> _______________________________________________
>>>> time-nuts mailing list -- time-nuts@lists.febo.com
>>>> To unsubscribe send an email to time-nuts-leave@lists.febo.com
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts@lists.febo.com
>>> To unsubscribe send an email to time-nuts-leave@lists.febo.com
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@lists.febo.com
>> To unsubscribe send an email to time-nuts-leave@lists.febo.com
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
R(
Richard (Rick) Karlquist
Fri, Mar 31, 2023 11:04 PM
I believe that IC went out of production decades ago.
Rick
On 3/31/2023 11:13 AM, Ed Marciniak via time-nuts wrote:
Linear technology made a thermoelectric full H bridge temperature controller. At the time, I’d like to say it came as a lot of two components. The second was an op-amp that may have been selected or trimmed unless there was a good reason to include it with samples (maybe less commonly available).
I believe that IC went out of production decades ago.
Rick
On 3/31/2023 11:13 AM, Ed Marciniak via time-nuts wrote:
> Linear technology made a thermoelectric full H bridge temperature controller. At the time, I’d like to say it came as a lot of two components. The second was an op-amp that may have been selected or trimmed unless there was a good reason to include it with samples (maybe less commonly available).
LJ
Lux, Jim
Fri, Mar 31, 2023 11:27 PM
On 3/31/23 4:04 PM, Richard (Rick) Karlquist via time-nuts wrote:
I believe that IC went out of production decades ago.
Rick
On 3/31/2023 11:13 AM, Ed Marciniak via time-nuts wrote:
Linear technology made a thermoelectric full H bridge temperature
controller. At the time, I’d like to say it came as a lot of two
components. The second was an op-amp that may have been selected or
trimmed unless there was a good reason to include it with samples
(maybe less commonly available).
On 3/31/23 4:04 PM, Richard (Rick) Karlquist via time-nuts wrote:
> I believe that IC went out of production decades ago.
>
> Rick
>
https://www.analog.com/en/products/ltc1923.html#product-overview
High Efficiency Thermoelectric Cooler Controller
Example circuit is 0.01degree.
Eval boards available too DC388C and DC491A
Looks like a $40 part, with a $150 eval board.
> On 3/31/2023 11:13 AM, Ed Marciniak via time-nuts wrote:
>> Linear technology made a thermoelectric full H bridge temperature
>> controller. At the time, I’d like to say it came as a lot of two
>> components. The second was an op-amp that may have been selected or
>> trimmed unless there was a good reason to include it with samples
>> (maybe less commonly available).
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
>
EM
Ed Marciniak
Sat, Apr 1, 2023 2:22 AM
Yes, that’s the one. A bit expensive, but current production.
My guess is it's probably used for DWDM lasers in telecom.
That said, call your local office and ask for a couple samples.
I still think it would make a fine way to use AT cut crystals instead of SC because it will no longer matter if the crystal you have happens to have a turnover above or below rooom temperature....and there are two turnover temperatures...you could use either of above ambient is no longer a requirement.
Get Outlook for iOShttps://aka.ms/o0ukef
From: Lux, Jim via time-nuts time-nuts@lists.febo.com
Sent: Friday, March 31, 2023 6:27:59 PM
To: time-nuts@lists.febo.com time-nuts@lists.febo.com
Cc: Lux, Jim jim@luxfamily.com
Subject: [time-nuts] Re: Temperature accuracy / repeatability / drift of double ovens
On 3/31/23 4:04 PM, Richard (Rick) Karlquist via time-nuts wrote:
I believe that IC went out of production decades ago.
Rick
On 3/31/2023 11:13 AM, Ed Marciniak via time-nuts wrote:
Linear technology made a thermoelectric full H bridge temperature
controller. At the time, I’d like to say it came as a lot of two
components. The second was an op-amp that may have been selected or
trimmed unless there was a good reason to include it with samples
(maybe less commonly available).
Yes, that’s the one. A bit expensive, but current production.
My guess is it's probably used for DWDM lasers in telecom.
That said, call your local office and ask for a couple samples.
I still think it would make a fine way to use AT cut crystals instead of SC because it will no longer matter if the crystal you have happens to have a turnover above or below rooom temperature....and there are two turnover temperatures...you could use either of above ambient is no longer a requirement.
Get Outlook for iOS<https://aka.ms/o0ukef>
________________________________
From: Lux, Jim via time-nuts <time-nuts@lists.febo.com>
Sent: Friday, March 31, 2023 6:27:59 PM
To: time-nuts@lists.febo.com <time-nuts@lists.febo.com>
Cc: Lux, Jim <jim@luxfamily.com>
Subject: [time-nuts] Re: Temperature accuracy / repeatability / drift of double ovens
On 3/31/23 4:04 PM, Richard (Rick) Karlquist via time-nuts wrote:
> I believe that IC went out of production decades ago.
>
> Rick
>
https://www.analog.com/en/products/ltc1923.html#product-overview
High Efficiency Thermoelectric Cooler Controller
Example circuit is 0.01degree.
Eval boards available too DC388C and DC491A
Looks like a $40 part, with a $150 eval board.
> On 3/31/2023 11:13 AM, Ed Marciniak via time-nuts wrote:
>> Linear technology made a thermoelectric full H bridge temperature
>> controller. At the time, I’d like to say it came as a lot of two
>> components. The second was an op-amp that may have been selected or
>> trimmed unless there was a good reason to include it with samples
>> (maybe less commonly available).
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
>
_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
G
ghf (at) hoffmann-hochfrequenz.de
Sat, Apr 1, 2023 9:12 PM
Am 2023-03-31 20:13, schrieb Ed Marciniak via time-nuts:
Linear technology made a thermoelectric full H bridge temperature
controller. At the time, I’d like to say it came as a lot of two
components. The second was an op-amp that may have been selected or
trimmed unless there was a good reason to include it with samples
(maybe less commonly available).
It was supposed to be able to achieve 10 milliKelvin stability easily,
and with more care in design, 1 milliKelvin stability was supposed to
be not too difficult to achieve.
That really gives me some hope, I'll take the care :-)
The question is now how to prove it!
It was a moderately expensive part…about $20 for the two components. A
handful of passives and a a pair of N-channel and a pair of P-channel
MOSFETs were required.
If memory serves correctly it was also switched mode rather than
linear.
Thanks for the pointer, also to Jim Lux.
As a first try, I used the Analog Devices ADN 8834, also from AD, now
that
they have bought LT. But the Linear Technology LTC1923 data sheet is in
a different league, they even give absolute numbers and strip charts.
I'd bet that this went over the desk of Jim Williams, RIP. Simply top
notch.
The ADN8843 has the H-Bridge built-in and it worked for someone I know.
I'm not too excited by the cascaded TECs. They can only provide a
temperature difference, and absolute temperature only with a regulation
loop with more time constants between the TEC and the NTC. The low
thermal
impedance path through the TEC is a real drawback, and probably the
culprit for the low efficiency. No wonder that no one uses them in
OCXOs.
But 87°C like in my Morion OCXOs is no solution, either, even if we
could
dim down the absolute temperature.
Thermal gain on a single oven is likely in the 300 to 600 range.
Double oven boosts that by 10 to 30X
Thermal gain in this case: take the ambient change divide it by the
thermal gain
and you get the oven temperature change.
Long term stability:
low enough that it does not impact aging. ( = that’s how you test it).
Since that gets back
to the specific thermistor being used, you are into the ‘likely not
much help” range.
Best guess info: The crystal + circuit is < 1x10^-9 / C in the
vicinity of the turn. The operating
point is offset from turn to minimize the combination of crystal +
circuit. Just what the net
is ….
No crystal today, just VCSELs...
Still, if you can “see” aging at the parts in the (low?) 10^-11 range,
it sort of kind of gets
you to a number. Maybe 0.01C / day.
Am 2023-03-31 20:13, schrieb Ed Marciniak via time-nuts:
> Linear technology made a thermoelectric full H bridge temperature
> controller. At the time, I’d like to say it came as a lot of two
> components. The second was an op-amp that may have been selected or
> trimmed unless there was a good reason to include it with samples
> (maybe less commonly available).
>
> It was supposed to be able to achieve 10 milliKelvin stability easily,
> and with more care in design, 1 milliKelvin stability was supposed to
> be not too difficult to achieve.
That really gives me some hope, I'll take the care :-)
The question is now how to prove it!
> It was a moderately expensive part…about $20 for the two components. A
> handful of passives and a a pair of N-channel and a pair of P-channel
> MOSFETs were required.
>
> If memory serves correctly it was also switched mode rather than
> linear.
Thanks for the pointer, also to Jim Lux.
As a first try, I used the Analog Devices ADN 8834, also from AD, now
that
they have bought LT. But the Linear Technology LTC1923 data sheet is in
a different league, they even give absolute numbers and strip charts.
I'd bet that this went over the desk of Jim Williams, RIP. Simply top
notch.
The ADN8843 has the H-Bridge built-in and it worked for someone I know.
I'm not too excited by the cascaded TECs. They can only provide a
temperature difference, and absolute temperature only with a regulation
loop with more time constants between the TEC and the NTC. The low
thermal
impedance path through the TEC is a real drawback, and probably the
culprit for the low efficiency. No wonder that no one uses them in
OCXOs.
But 87°C like in my Morion OCXOs is no solution, either, even if we
could
dim down the absolute temperature.
> From: Bob Camp via time-nuts <time-nuts@lists.febo.com>
> Sent: Thursday, March 30, 2023 6:01:55 PM
> Thermal gain on a single oven is likely in the 300 to 600 range.
>
> Double oven boosts that by 10 to 30X
>
> Thermal gain in this case: take the ambient change divide it by the
> thermal gain
> and you get the oven temperature change.
>
> Long term stability:
>
> low enough that it does not impact aging. ( = that’s how you test it).
> Since that gets back
> to the specific thermistor being used, you are into the ‘likely not
> much help” range.
>
> Best guess info: The crystal + circuit is < 1x10^-9 / C in the
> vicinity of the turn. The operating
> point is offset from turn to minimize the combination of crystal +
> circuit. Just what the net
> is ….
No crystal today, just VCSELs...
> Still, if you can “see” aging at the parts in the (low?) 10^-11 range,
> it sort of kind of gets
> you to a number. Maybe 0.01C / day.
>> Cheers, Gerhard
BC
Bob Camp
Sat, Apr 1, 2023 10:29 PM
Hi
They tried that (cooling vs heating) back in the 1950’s. It created more issues
(heat sinking the lossy TEC’s, poor TEC reliability, ….. ) than it fixed.
Bob
On Mar 31, 2023, at 10:22 PM, Ed Marciniak via time-nuts time-nuts@lists.febo.com wrote:
Yes, that’s the one. A bit expensive, but current production.
My guess is it's probably used for DWDM lasers in telecom.
That said, call your local office and ask for a couple samples.
I still think it would make a fine way to use AT cut crystals instead of SC because it will no longer matter if the crystal you have happens to have a turnover above or below rooom temperature....and there are two turnover temperatures...you could use either of above ambient is no longer a requirement.
Get Outlook for iOShttps://aka.ms/o0ukef
From: Lux, Jim via time-nuts time-nuts@lists.febo.com
Sent: Friday, March 31, 2023 6:27:59 PM
To: time-nuts@lists.febo.com time-nuts@lists.febo.com
Cc: Lux, Jim jim@luxfamily.com
Subject: [time-nuts] Re: Temperature accuracy / repeatability / drift of double ovens
On 3/31/23 4:04 PM, Richard (Rick) Karlquist via time-nuts wrote:
I believe that IC went out of production decades ago.
Rick
On 3/31/2023 11:13 AM, Ed Marciniak via time-nuts wrote:
Linear technology made a thermoelectric full H bridge temperature
controller. At the time, I’d like to say it came as a lot of two
components. The second was an op-amp that may have been selected or
trimmed unless there was a good reason to include it with samples
(maybe less commonly available).
Hi
They tried that (cooling vs heating) back in the 1950’s. It created more issues
(heat sinking the lossy TEC’s, poor TEC reliability, ….. ) than it fixed.
Bob
> On Mar 31, 2023, at 10:22 PM, Ed Marciniak via time-nuts <time-nuts@lists.febo.com> wrote:
>
> Yes, that’s the one. A bit expensive, but current production.
>
> My guess is it's probably used for DWDM lasers in telecom.
>
> That said, call your local office and ask for a couple samples.
>
> I still think it would make a fine way to use AT cut crystals instead of SC because it will no longer matter if the crystal you have happens to have a turnover above or below rooom temperature....and there are two turnover temperatures...you could use either of above ambient is no longer a requirement.
>
> Get Outlook for iOS<https://aka.ms/o0ukef>
> ________________________________
> From: Lux, Jim via time-nuts <time-nuts@lists.febo.com>
> Sent: Friday, March 31, 2023 6:27:59 PM
> To: time-nuts@lists.febo.com <time-nuts@lists.febo.com>
> Cc: Lux, Jim <jim@luxfamily.com>
> Subject: [time-nuts] Re: Temperature accuracy / repeatability / drift of double ovens
>
> On 3/31/23 4:04 PM, Richard (Rick) Karlquist via time-nuts wrote:
>> I believe that IC went out of production decades ago.
>>
>> Rick
>>
> https://www.analog.com/en/products/ltc1923.html#product-overview
>
>
> High Efficiency Thermoelectric Cooler Controller
>
> Example circuit is 0.01degree.
>
> Eval boards available too DC388C and DC491A
>
>
> Looks like a $40 part, with a $150 eval board.
>
>
>
>
>
>
>> On 3/31/2023 11:13 AM, Ed Marciniak via time-nuts wrote:
>>> Linear technology made a thermoelectric full H bridge temperature
>>> controller. At the time, I’d like to say it came as a lot of two
>>> components. The second was an op-amp that may have been selected or
>>> trimmed unless there was a good reason to include it with samples
>>> (maybe less commonly available).
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@lists.febo.com
>> To unsubscribe send an email to time-nuts-leave@lists.febo.com
>>
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
JH
Javier Herrero
Sun, Apr 2, 2023 8:42 PM
Am 2023-03-31 20:13, schrieb Ed Marciniak via time-nuts:
Linear technology made a thermoelectric full H bridge temperature
controller. At the time, I’d like to say it came as a lot of two
components. The second was an op-amp that may have been selected or
trimmed unless there was a good reason to include it with samples
(maybe less commonly available).
It was supposed to be able to achieve 10 milliKelvin stability easily,
and with more care in design, 1 milliKelvin stability was supposed to
be not too difficult to achieve.
That really gives me some hope, I'll take the care :-)
The question is now how to prove it!
It was a moderately expensive part…about $20 for the two components. A
handful of passives and a a pair of N-channel and a pair of P-channel
MOSFETs were required.
If memory serves correctly it was also switched mode rather than linear.
Thanks for the pointer, also to Jim Lux.
As a first try, I used the Analog Devices ADN 8834, also from AD, now
that
they have bought LT. But the Linear Technology LTC1923 data sheet is in
a different league, they even give absolute numbers and strip charts.
I'd bet that this went over the desk of Jim Williams, RIP. Simply top
notch.
The ADN8843 has the H-Bridge built-in and it worked for someone I know.
I'm not too excited by the cascaded TECs. They can only provide a
temperature difference, and absolute temperature only with a regulation
loop with more time constants between the TEC and the NTC. The low
thermal
impedance path through the TEC is a real drawback, and probably the
culprit for the low efficiency. No wonder that no one uses them in OCXOs.
But 87°C like in my Morion OCXOs is no solution, either, even if we could
dim down the absolute temperature.
Thermal gain on a single oven is likely in the 300 to 600 range.
Double oven boosts that by 10 to 30X
Thermal gain in this case: take the ambient change divide it by the
thermal gain
and you get the oven temperature change.
Long term stability:
low enough that it does not impact aging. ( = that’s how you test it).
Since that gets back
to the specific thermistor being used, you are into the ‘likely not
much help” range.
Best guess info: The crystal + circuit is < 1x10^-9 / C in the
vicinity of the turn. The operating
point is offset from turn to minimize the combination of crystal +
circuit. Just what the net
is ….
No crystal today, just VCSELs...
Still, if you can “see” aging at the parts in the (low?) 10^-11 range,
it sort of kind of gets
you to a number. Maybe 0.01C / day.
Hello,
Of course, the data sheet has Jim Williams signature. Take a look to his
app note
https://www.analog.com/media/en/technical-documentation/application-notes/an89.pdf
and don't miss the last page :)
Regards,
Javier, EA1CRB
On 4/1/23 23:11, Gerhard Hoffmann via time-nuts wrote:
> Am 2023-03-31 20:13, schrieb Ed Marciniak via time-nuts:
>> Linear technology made a thermoelectric full H bridge temperature
>> controller. At the time, I’d like to say it came as a lot of two
>> components. The second was an op-amp that may have been selected or
>> trimmed unless there was a good reason to include it with samples
>> (maybe less commonly available).
>>
>> It was supposed to be able to achieve 10 milliKelvin stability easily,
>> and with more care in design, 1 milliKelvin stability was supposed to
>> be not too difficult to achieve.
>
> That really gives me some hope, I'll take the care :-)
> The question is now how to prove it!
>
>
>> It was a moderately expensive part…about $20 for the two components. A
>> handful of passives and a a pair of N-channel and a pair of P-channel
>> MOSFETs were required.
>>
>> If memory serves correctly it was also switched mode rather than linear.
>
> Thanks for the pointer, also to Jim Lux.
>
> As a first try, I used the Analog Devices ADN 8834, also from AD, now
> that
> they have bought LT. But the Linear Technology LTC1923 data sheet is in
> a different league, they even give absolute numbers and strip charts.
> I'd bet that this went over the desk of Jim Williams, RIP. Simply top
> notch.
> The ADN8843 has the H-Bridge built-in and it worked for someone I know.
>
> I'm not too excited by the cascaded TECs. They can only provide a
> temperature difference, and absolute temperature only with a regulation
> loop with more time constants between the TEC and the NTC. The low
> thermal
> impedance path through the TEC is a real drawback, and probably the
> culprit for the low efficiency. No wonder that no one uses them in OCXOs.
> But 87°C like in my Morion OCXOs is no solution, either, even if we could
> dim down the absolute temperature.
>
>
>> From: Bob Camp via time-nuts <time-nuts@lists.febo.com>
>> Sent: Thursday, March 30, 2023 6:01:55 PM
>
>> Thermal gain on a single oven is likely in the 300 to 600 range.
>>
>> Double oven boosts that by 10 to 30X
>>
>> Thermal gain in this case: take the ambient change divide it by the
>> thermal gain
>> and you get the oven temperature change.
>>
>> Long term stability:
>>
>> low enough that it does not impact aging. ( = that’s how you test it).
>> Since that gets back
>> to the specific thermistor being used, you are into the ‘likely not
>> much help” range.
>>
>> Best guess info: The crystal + circuit is < 1x10^-9 / C in the
>> vicinity of the turn. The operating
>> point is offset from turn to minimize the combination of crystal +
>> circuit. Just what the net
>> is ….
>
> No crystal today, just VCSELs...
>
>> Still, if you can “see” aging at the parts in the (low?) 10^-11 range,
>> it sort of kind of gets
>> you to a number. Maybe 0.01C / day.
>
>
>>> Cheers, Gerhard
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe send an email to time-nuts-leave@lists.febo.com
LV
Lester Veenstra
Wed, Apr 5, 2023 3:40 PM
Also any devotees of Jim should not miss his foot note: Page 5 Note 6 of the data sheet
https://www.analog.com/media/en/technical-documentation/application-notes/an89.pdf
Lester B Veenstra K1YCM MØYCM W8YCM 6Y6Y W8YCM/6Y 6Y8LV (Reformed USNSG CTM1)
lester@veenstras.com
452 Stable Ln
Keyser WV 26726 USA
GPS: 39.336826 N 78.982287 W (Google)
GPS: 39.33682 N 78.9823741 W (GPSDO)
Telephones:
Home: +1-304-289-6057
US cell +1-304-790-9192
Jamaica cell: +1-876-456-8898
-----Original Message-----
From: Javier Herrero via time-nuts [mailto:time-nuts@lists.febo.com]
Sent: Sunday, April 2, 2023 4:42 PM
To: time-nuts@lists.febo.com
Cc: Javier Herrero
Subject: [time-nuts] Re: Temperature accuracy / repeatability / drift of double ovens
Hello,
Of course, the data sheet has Jim Williams signature. Take a look to his
app note
https://www.analog.com/media/en/technical-documentation/application-notes/an89.pdf
and don't miss the last page :)
Regards,
Javier, EA1CRB
Also any devotees of Jim should not miss his foot note: Page 5 Note 6 of the data sheet
https://www.analog.com/media/en/technical-documentation/application-notes/an89.pdf
Lester B Veenstra K1YCM MØYCM W8YCM 6Y6Y W8YCM/6Y 6Y8LV (Reformed USNSG CTM1)
lester@veenstras.com
452 Stable Ln
Keyser WV 26726 USA
GPS: 39.336826 N 78.982287 W (Google)
GPS: 39.33682 N 78.9823741 W (GPSDO)
Telephones:
Home: +1-304-289-6057
US cell +1-304-790-9192
Jamaica cell: +1-876-456-8898
-----Original Message-----
From: Javier Herrero via time-nuts [mailto:time-nuts@lists.febo.com]
Sent: Sunday, April 2, 2023 4:42 PM
To: time-nuts@lists.febo.com
Cc: Javier Herrero
Subject: [time-nuts] Re: Temperature accuracy / repeatability / drift of double ovens
Hello,
Of course, the data sheet has Jim Williams signature. Take a look to his
app note
https://www.analog.com/media/en/technical-documentation/application-notes/an89.pdf
and don't miss the last page :)
Regards,