time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

PRS-10 Warm-up Time, Calibrating/Adjusting, and long-term poweron

FC
Forrest Christian (List Account)
Fri, Mar 8, 2019 8:18 AM

Hopefully you'll all grace me with a few answers to a beginner
time-nut question or two.

I have a PRS-10 I've never used other than to power it on with a
recently-acquired heatsink and verify that it seems to operate
correctly and that the operational parameters don't seem out of
tolerance.  I would like to use this in the near future as a 10Mhz
reference for a TAPR TICC which I'd like to use to measure the jitter
performance of the PPS output of various consumer GPS receivers, the
goal being to end up with a jitter histogram.

So three interrelated questions:

  1. Assuming the PRS-10 has been off for a long time, how long should I
    plan on leaving this on for the 10Mhz to stabilize?  I see the
    longest warmup time on the spec-sheet is 7 minutes -  although this
    seems a lot shorter than I'd likely use in real life,  I'm also not
    sure if there's much benefit to an excessively longer warmup time
    (like days), would like opinions on this.

  2. Longer-term I'd like to use the 1PPS output from a Trimble
    Thunderbolt to calibrate the PRS10 and adjust if necessary just to
    trim out any aging drift on the PRS10.  Initially I thought I was
    going to discipline the PRS10 on a continual basis with the
    Thunderbolt using the PPS input on the PRS10, but I've recently
    realized that leaving the PRS10 on permanently might not be the best
    option (see Question 3).  So I'm looking for opinions on how to keep
    the PRS10 calibrated/adjusted.  I.E. trim with the trimmer, adjust
    using digital commands, etc.

  3. As implied in #2, I was originally planning on leaving the PRS10 on
    a continuous basis.  I've read a couple of things which imply that
    there is little benefit to doing so, and that every hour it's on
    consumes the lamp life.  Assuming I only need the highly stable PRS10
    source every few months for things like jitter measurements on 1PPS
    sources, is there any benefit to leaving the PRS10 on?

--

  • Forrest
Hopefully you'll all grace me with a few answers to a beginner time-nut question or two. I have a PRS-10 I've never used other than to power it on with a recently-acquired heatsink and verify that it seems to operate correctly and that the operational parameters don't seem out of tolerance. I would like to use this in the near future as a 10Mhz reference for a TAPR TICC which I'd like to use to measure the jitter performance of the PPS output of various consumer GPS receivers, the goal being to end up with a jitter histogram. So three interrelated questions: 1) Assuming the PRS-10 has been off for a long time, how long should I plan on leaving this on for the 10Mhz to stabilize? I see the longest warmup time on the spec-sheet is 7 minutes - although this seems a lot shorter than I'd likely use in real life, I'm also not sure if there's much benefit to an excessively longer warmup time (like days), would like opinions on this. 2) Longer-term I'd like to use the 1PPS output from a Trimble Thunderbolt to calibrate the PRS10 and adjust if necessary just to trim out any aging drift on the PRS10. Initially I thought I was going to discipline the PRS10 on a continual basis with the Thunderbolt using the PPS input on the PRS10, but I've recently realized that leaving the PRS10 on permanently might not be the best option (see Question 3). So I'm looking for opinions on how to keep the PRS10 calibrated/adjusted. I.E. trim with the trimmer, adjust using digital commands, etc. 3) As implied in #2, I was originally planning on leaving the PRS10 on a continuous basis. I've read a couple of things which imply that there is little benefit to doing so, and that every hour it's on consumes the lamp life. Assuming I only need the highly stable PRS10 source every few months for things like jitter measurements on 1PPS sources, is there any benefit to leaving the PRS10 on? -- - Forrest
DW
Dana Whitlow
Fri, Mar 8, 2019 12:31 PM

Keep in mind that the 7 min "warmup time" of the PRS-10 is jut the time for
the
unit to get the physics package close enough to final temperature for the
dithering
loop to lock to the atomic transition. But based on measurements of current
drawn
by the oven(s), it takes more like an hour for the ovens to settle down
very well.

My own experiments with a well-used "telecom mod" PRS-10 suggest that it
takes
my unit more like 3 or 4 days to really settle down to its background
frequency drift
rate.  This rate is specified as something like "< 5E-11 per month", which
is frankly
pretty awful.  Mine seems to be better than spec, more like 2E-11 per month
(frequency
increasing over time).  The point here is that there are apparently a
number of warm
up drift mechanisms operating, some of which take days to sensibly settle
down.

I've been wondering about the issue of whether it's best to operate a
PRS-10 only
at times of need, or continuously, with respect to life.  I posed the
question direct;y
to SRS, and they claim that it is better to leave it running all the time.
My own
"algorithm" has become "leave it on all the time" when I'm around and even
remotely
active with time-nuttery, but shut it down when I'm away on vacation.  This
is mostly
out of concern for something going wrong with the power supply and burning
the
house down.  Lightning storms are one such risk, for example.

Dana    K8YUM

On Fri, Mar 8, 2019 at 3:00 AM Forrest Christian (List Account) <
lists@packetflux.com> wrote:

Hopefully you'll all grace me with a few answers to a beginner
time-nut question or two.

I have a PRS-10 I've never used other than to power it on with a
recently-acquired heatsink and verify that it seems to operate
correctly and that the operational parameters don't seem out of
tolerance.  I would like to use this in the near future as a 10Mhz
reference for a TAPR TICC which I'd like to use to measure the jitter
performance of the PPS output of various consumer GPS receivers, the
goal being to end up with a jitter histogram.

So three interrelated questions:

  1. Assuming the PRS-10 has been off for a long time, how long should I
    plan on leaving this on for the 10Mhz to stabilize?  I see the
    longest warmup time on the spec-sheet is 7 minutes -  although this
    seems a lot shorter than I'd likely use in real life,  I'm also not
    sure if there's much benefit to an excessively longer warmup time
    (like days), would like opinions on this.

  2. Longer-term I'd like to use the 1PPS output from a Trimble
    Thunderbolt to calibrate the PRS10 and adjust if necessary just to
    trim out any aging drift on the PRS10.  Initially I thought I was
    going to discipline the PRS10 on a continual basis with the
    Thunderbolt using the PPS input on the PRS10, but I've recently
    realized that leaving the PRS10 on permanently might not be the best
    option (see Question 3).  So I'm looking for opinions on how to keep
    the PRS10 calibrated/adjusted.  I.E. trim with the trimmer, adjust
    using digital commands, etc.

  3. As implied in #2, I was originally planning on leaving the PRS10 on
    a continuous basis.  I've read a couple of things which imply that
    there is little benefit to doing so, and that every hour it's on
    consumes the lamp life.  Assuming I only need the highly stable PRS10
    source every few months for things like jitter measurements on 1PPS
    sources, is there any benefit to leaving the PRS10 on?

--

  • Forrest

time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Keep in mind that the 7 min "warmup time" of the PRS-10 is jut the time for the unit to get the physics package close enough to final temperature for the dithering loop to lock to the atomic transition. But based on measurements of current drawn by the oven(s), it takes more like an hour for the ovens to settle down very well. My own experiments with a well-used "telecom mod" PRS-10 suggest that it takes my unit more like 3 or 4 days to really settle down to its background frequency drift rate. This rate is specified as something like "< 5E-11 per month", which is frankly pretty awful. Mine seems to be better than spec, more like 2E-11 per month (frequency increasing over time). The point here is that there are apparently a number of warm up drift mechanisms operating, some of which take days to sensibly settle down. I've been wondering about the issue of whether it's best to operate a PRS-10 only at times of need, or continuously, with respect to life. I posed the question direct;y to SRS, and they claim that it is better to leave it running all the time. My own "algorithm" has become "leave it on all the time" when I'm around and even remotely active with time-nuttery, but shut it down when I'm away on vacation. This is mostly out of concern for something going wrong with the power supply and burning the house down. Lightning storms are one such risk, for example. Dana K8YUM On Fri, Mar 8, 2019 at 3:00 AM Forrest Christian (List Account) < lists@packetflux.com> wrote: > Hopefully you'll all grace me with a few answers to a beginner > time-nut question or two. > > I have a PRS-10 I've never used other than to power it on with a > recently-acquired heatsink and verify that it seems to operate > correctly and that the operational parameters don't seem out of > tolerance. I would like to use this in the near future as a 10Mhz > reference for a TAPR TICC which I'd like to use to measure the jitter > performance of the PPS output of various consumer GPS receivers, the > goal being to end up with a jitter histogram. > > So three interrelated questions: > > 1) Assuming the PRS-10 has been off for a long time, how long should I > plan on leaving this on for the 10Mhz to stabilize? I see the > longest warmup time on the spec-sheet is 7 minutes - although this > seems a lot shorter than I'd likely use in real life, I'm also not > sure if there's much benefit to an excessively longer warmup time > (like days), would like opinions on this. > > 2) Longer-term I'd like to use the 1PPS output from a Trimble > Thunderbolt to calibrate the PRS10 and adjust if necessary just to > trim out any aging drift on the PRS10. Initially I thought I was > going to discipline the PRS10 on a continual basis with the > Thunderbolt using the PPS input on the PRS10, but I've recently > realized that leaving the PRS10 on permanently might not be the best > option (see Question 3). So I'm looking for opinions on how to keep > the PRS10 calibrated/adjusted. I.E. trim with the trimmer, adjust > using digital commands, etc. > > 3) As implied in #2, I was originally planning on leaving the PRS10 on > a continuous basis. I've read a couple of things which imply that > there is little benefit to doing so, and that every hour it's on > consumes the lamp life. Assuming I only need the highly stable PRS10 > source every few months for things like jitter measurements on 1PPS > sources, is there any benefit to leaving the PRS10 on? > > -- > - Forrest > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >
AK
Attila Kinali
Fri, Mar 8, 2019 12:47 PM

On Fri, 8 Mar 2019 01:18:45 -0700
"Forrest Christian (List Account)" lists@packetflux.com wrote:

  1. Assuming the PRS-10 has been off for a long time, how long should I
    plan on leaving this on for the 10Mhz to stabilize?  I see the
    longest warmup time on the spec-sheet is 7 minutes -  although this
    seems a lot shorter than I'd likely use in real life,  I'm also not
    sure if there's much benefit to an excessively longer warmup time
    (like days), would like opinions on this.

The warm-up time is from "I switched it on" to "there is a 10MHz signal
and it's 'stable'". I.e. at latest after 7 minutes you should be able
to use the ouptut of the PRS10. But keep in mind that the stable-after-warm-up
is not tha same as stable-for-a-time-nut. A Rb vapor cell standard has
a retrace (ie change of frequency after being switched on, for details see
John Vig's tutorial) of several weeks to a few months, depending on the
exact construction of the Rb cell, how long it has been in use and how
long it has been off.

  1. Longer-term I'd like to use the 1PPS output from a Trimble
    Thunderbolt to calibrate the PRS10 and adjust if necessary just to
    trim out any aging drift on the PRS10.  Initially I thought I was
    going to discipline the PRS10 on a continual basis with the
    Thunderbolt using the PPS input on the PRS10, but I've recently
    realized that leaving the PRS10 on permanently might not be the best
    option (see Question 3).  So I'm looking for opinions on how to keep
    the PRS10 calibrated/adjusted.  I.E. trim with the trimmer, adjust
    using digital commands, etc.

Switching it on and off several times a month would be worse in
terms of longevity  than having it running continuously. The vapor
cell itself is usually run between 60°C and 100°C and the Rb lamp
usually between 120°C and 170°C. As you can imagine, power cycling
gives a huge thermal strain on all the components. The inner construction
of the PRS10 is not optimized for power cycling but thermal insulation
of the sensitive components (see e.g. http://time.kinali.ch/Rb/PRS10/ )

So I would rather feed the PPS to the PRS10 directly and let itself
do the work. This will also give you an optimized control loop for
the performance of the PRS10 without you having to go through the
process of desiging and optimizing the loop.

  1. As implied in #2, I was originally planning on leaving the PRS10 on
    a continuous basis.  I've read a couple of things which imply that
    there is little benefit to doing so, and that every hour it's on
    consumes the lamp life.  Assuming I only need the highly stable PRS10
    source every few months for things like jitter measurements on 1PPS
    sources, is there any benefit to leaving the PRS10 on?

If you want to do just jitter measurement, then I wouldn't use a Rb standard
but a stable OCXO instead. At averaging times below 1-10s an OCXO will beat
any Rb standard, even one that has an OCXO like the PRS10 (admittedly, it's
not a high quality OCXO but it's one none-the-less). An OCXO has a shorter
warm-up time (usually 2-3 minutes until stable, usually less than a day
until time-nuts-stable, less than a month until nutty-time-nut-stable).

I personally, would use an OCXO as reference to the TICC (or TICCs?)
and use the PRS10 as an additional input to the TICC. This way you
can do a multi-way comparison between the GPSDOs and the Rb, removing
out the (in-)stability of the reference OCXO.

Beside those comments, you have not told us at what kind of performance
you are aiming at. There is a huge difference whether you need 1e-10 or
1e-13 and whether you need it at 1s, 1ks or even 100ks averaging times.

			Attila Kinali

--
It is upon moral qualities that a society is ultimately founded. All
the prosperity and technological sophistication in the world is of no
use without that foundation.
-- Miss Matheson, The Diamond Age, Neal Stephenson

On Fri, 8 Mar 2019 01:18:45 -0700 "Forrest Christian (List Account)" <lists@packetflux.com> wrote: > 1) Assuming the PRS-10 has been off for a long time, how long should I > plan on leaving this on for the 10Mhz to stabilize? I see the > longest warmup time on the spec-sheet is 7 minutes - although this > seems a lot shorter than I'd likely use in real life, I'm also not > sure if there's much benefit to an excessively longer warmup time > (like days), would like opinions on this. The warm-up time is from "I switched it on" to "there is a 10MHz signal and it's 'stable'". I.e. at latest after 7 minutes you should be able to use the ouptut of the PRS10. But keep in mind that the stable-after-warm-up is not tha same as stable-for-a-time-nut. A Rb vapor cell standard has a retrace (ie change of frequency after being switched on, for details see John Vig's tutorial) of several weeks to a few months, depending on the exact construction of the Rb cell, how long it has been in use and how long it has been off. > 2) Longer-term I'd like to use the 1PPS output from a Trimble > Thunderbolt to calibrate the PRS10 and adjust if necessary just to > trim out any aging drift on the PRS10. Initially I thought I was > going to discipline the PRS10 on a continual basis with the > Thunderbolt using the PPS input on the PRS10, but I've recently > realized that leaving the PRS10 on permanently might not be the best > option (see Question 3). So I'm looking for opinions on how to keep > the PRS10 calibrated/adjusted. I.E. trim with the trimmer, adjust > using digital commands, etc. Switching it on and off several times a month would be worse in terms of longevity than having it running continuously. The vapor cell itself is usually run between 60°C and 100°C and the Rb lamp usually between 120°C and 170°C. As you can imagine, power cycling gives a huge thermal strain on all the components. The inner construction of the PRS10 is not optimized for power cycling but thermal insulation of the sensitive components (see e.g. http://time.kinali.ch/Rb/PRS10/ ) So I would rather feed the PPS to the PRS10 directly and let itself do the work. This will also give you an optimized control loop for the performance of the PRS10 without you having to go through the process of desiging and optimizing the loop. > 3) As implied in #2, I was originally planning on leaving the PRS10 on > a continuous basis. I've read a couple of things which imply that > there is little benefit to doing so, and that every hour it's on > consumes the lamp life. Assuming I only need the highly stable PRS10 > source every few months for things like jitter measurements on 1PPS > sources, is there any benefit to leaving the PRS10 on? If you want to do just jitter measurement, then I wouldn't use a Rb standard but a stable OCXO instead. At averaging times below 1-10s an OCXO will beat any Rb standard, even one that has an OCXO like the PRS10 (admittedly, it's not a high quality OCXO but it's one none-the-less). An OCXO has a shorter warm-up time (usually 2-3 minutes until stable, usually less than a day until time-nuts-stable, less than a month until nutty-time-nut-stable). I personally, would use an OCXO as reference to the TICC (or TICCs?) and use the PRS10 as an additional input to the TICC. This way you can do a multi-way comparison between the GPSDOs and the Rb, removing out the (in-)stability of the reference OCXO. Beside those comments, you have not told us at what kind of performance you are aiming at. There is a huge difference whether you need 1e-10 or 1e-13 and whether you need it at 1s, 1ks or even 100ks averaging times. Attila Kinali -- It is upon moral qualities that a society is ultimately founded. All the prosperity and technological sophistication in the world is of no use without that foundation. -- Miss Matheson, The Diamond Age, Neal Stephenson
BK
Bob kb8tq
Fri, Mar 8, 2019 1:28 PM

Hi

Some of this gets into just what you are trying to do. Cycle to cycle jitter on the
PPS output of most GPS modules is in the “many ns” range. If sawtooth correction
is not being applied, a measurement that is good to an accuracy of 1 ns / 5 = 200 ps
would be plenty good enough. That is a 2x10^-10 (0.2 ppb) sort of accuracy.

If you are looking at longer term effects, you may need better accuracy / lower drift.
You can easily get past what a nice new 5071 can do heading down this road … For
now let’s assume the target above is correct.

The PRS-10 should be within 0.2 ppb within a few minutes after locking up. It should
hold this sort of accuracy “on the shelf / power off” pretty much forever and ever. In fact,
it’s a pretty good bet that it will hold 10X this accuracy for quite a long time. You probably
should warm it up for a day or so if you are after 0.02 ppb.

Calibration wise, the “easy” way to do it depends a lot on what you have to do the measurement.
If you have a TICC (sounds like you do) - use it. You can also use a scope or a more
conventional counter.

If you are going after 0.02 ppb, you would want to set the unit to 0.002 ppb. That gives you
90% of the budget for things like temperature drift and aging. Indeed you might want to go
to a tighter number …. this is Time Nuts after all.

0.002 ppb is 2 ns over 1,000 seconds. It’s 200 ns over 100,000 seconds. (Amazing how that
works :) ). Is your TBolt PPS good to 2 ns? Probably not. Is it good to 200 ns? most certainly.
Since 100,000 seconds is just over a day, that makes for a fairly easy adjustment process.
Look at it once a day and tweak it. After a while it will not need tweaks any more.

The advantage of a “long term” adjustment like this is that it also takes in the temperature
swings in your lab over a one day period. At the 0.002 ppb level, they will be the dominant
part of what you see. Will you get to the 0.002 ppb level? It depends a lot on just how stable
your particular unit is and how drafty your lab is.

Running any sort of electronics gear 24 hours a day is a risk. It may wear out, it may cause
a fire, it may be fine … who knows. It also pulls power and you pay the electric company for that.
It’s generally cheaper / safer to turn stuff off if it is only going to be used rarely.

Lots of fun !!!

Bob

On Mar 8, 2019, at 3:18 AM, Forrest Christian (List Account) lists@packetflux.com wrote:

Hopefully you'll all grace me with a few answers to a beginner
time-nut question or two.

I have a PRS-10 I've never used other than to power it on with a
recently-acquired heatsink and verify that it seems to operate
correctly and that the operational parameters don't seem out of
tolerance.  I would like to use this in the near future as a 10Mhz
reference for a TAPR TICC which I'd like to use to measure the jitter
performance of the PPS output of various consumer GPS receivers, the
goal being to end up with a jitter histogram.

So three interrelated questions:

  1. Assuming the PRS-10 has been off for a long time, how long should I
    plan on leaving this on for the 10Mhz to stabilize?  I see the
    longest warmup time on the spec-sheet is 7 minutes -  although this
    seems a lot shorter than I'd likely use in real life,  I'm also not
    sure if there's much benefit to an excessively longer warmup time
    (like days), would like opinions on this.

  2. Longer-term I'd like to use the 1PPS output from a Trimble
    Thunderbolt to calibrate the PRS10 and adjust if necessary just to
    trim out any aging drift on the PRS10.  Initially I thought I was
    going to discipline the PRS10 on a continual basis with the
    Thunderbolt using the PPS input on the PRS10, but I've recently
    realized that leaving the PRS10 on permanently might not be the best
    option (see Question 3).  So I'm looking for opinions on how to keep
    the PRS10 calibrated/adjusted.  I.E. trim with the trimmer, adjust
    using digital commands, etc.

  3. As implied in #2, I was originally planning on leaving the PRS10 on
    a continuous basis.  I've read a couple of things which imply that
    there is little benefit to doing so, and that every hour it's on
    consumes the lamp life.  Assuming I only need the highly stable PRS10
    source every few months for things like jitter measurements on 1PPS
    sources, is there any benefit to leaving the PRS10 on?

--

  • Forrest

time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi Some of this gets into just what you are trying to do. Cycle to cycle jitter on the PPS output of most GPS modules is in the “many ns” range. If sawtooth correction is not being applied, a measurement that is good to an accuracy of 1 ns / 5 = 200 ps would be plenty good enough. That is a 2x10^-10 (0.2 ppb) sort of accuracy. If you are looking at longer term effects, you may need better accuracy / lower drift. You can easily get past what a nice new 5071 can do heading down this road … For now let’s assume the target above is correct. The PRS-10 should be within 0.2 ppb within a few minutes after locking up. It should hold this sort of accuracy “on the shelf / power off” pretty much forever and ever. In fact, it’s a pretty good bet that it will hold 10X this accuracy for quite a long time. You probably should warm it up for a day or so if you are after 0.02 ppb. Calibration wise, the “easy” way to do it depends a lot on what you have to do the measurement. If you have a TICC (sounds like you do) - use it. You can also use a scope or a more conventional counter. If you are going after 0.02 ppb, you would want to set the unit to 0.002 ppb. That gives you 90% of the budget for things like temperature drift and aging. Indeed you might want to go to a tighter number …. this is Time Nuts after all. 0.002 ppb is 2 ns over 1,000 seconds. It’s 200 ns over 100,000 seconds. (Amazing how that works :) ). Is your TBolt PPS good to 2 ns? Probably not. Is it good to 200 ns? most certainly. Since 100,000 seconds is just over a day, that makes for a fairly easy adjustment process. Look at it once a day and tweak it. After a while it will not need tweaks any more. The advantage of a “long term” adjustment like this is that it also takes in the temperature swings in your lab over a one day period. At the 0.002 ppb level, they will be the dominant part of what you see. Will you get to the 0.002 ppb level? It depends a lot on just how stable your particular unit is and how drafty your lab is. Running any sort of electronics gear 24 hours a day is a risk. It may wear out, it may cause a fire, it may be fine … who knows. It also pulls power and you pay the electric company for that. It’s generally cheaper / safer to turn stuff off if it is only going to be used rarely. Lots of fun !!! Bob > On Mar 8, 2019, at 3:18 AM, Forrest Christian (List Account) <lists@packetflux.com> wrote: > > Hopefully you'll all grace me with a few answers to a beginner > time-nut question or two. > > I have a PRS-10 I've never used other than to power it on with a > recently-acquired heatsink and verify that it seems to operate > correctly and that the operational parameters don't seem out of > tolerance. I would like to use this in the near future as a 10Mhz > reference for a TAPR TICC which I'd like to use to measure the jitter > performance of the PPS output of various consumer GPS receivers, the > goal being to end up with a jitter histogram. > > So three interrelated questions: > > 1) Assuming the PRS-10 has been off for a long time, how long should I > plan on leaving this on for the 10Mhz to stabilize? I see the > longest warmup time on the spec-sheet is 7 minutes - although this > seems a lot shorter than I'd likely use in real life, I'm also not > sure if there's much benefit to an excessively longer warmup time > (like days), would like opinions on this. > > 2) Longer-term I'd like to use the 1PPS output from a Trimble > Thunderbolt to calibrate the PRS10 and adjust if necessary just to > trim out any aging drift on the PRS10. Initially I thought I was > going to discipline the PRS10 on a continual basis with the > Thunderbolt using the PPS input on the PRS10, but I've recently > realized that leaving the PRS10 on permanently might not be the best > option (see Question 3). So I'm looking for opinions on how to keep > the PRS10 calibrated/adjusted. I.E. trim with the trimmer, adjust > using digital commands, etc. > > 3) As implied in #2, I was originally planning on leaving the PRS10 on > a continuous basis. I've read a couple of things which imply that > there is little benefit to doing so, and that every hour it's on > consumes the lamp life. Assuming I only need the highly stable PRS10 > source every few months for things like jitter measurements on 1PPS > sources, is there any benefit to leaving the PRS10 on? > > -- > - Forrest > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
TK
Taka Kamiya
Fri, Mar 8, 2019 2:09 PM

I am experimenting with exactly the same thing.  PRS-10 with Thunderbolt. 

PRS-10 takes about a day to stabilize.  When physics lock happens, it has limited stability - which is mentioned in manual.  Before I calibrated, I waited for 2 days.  These movements are visible if you feed GPSDO into channel 1 of scope and set a trigger on it, and feed PRS-10 output to channel 2.
For thunderbolt GPSDO, I am using a telecom unit, Nortel GPSTM modified with 1 pps output.  PRS-10 can take this input and lock, but it takes 1 to 2  hours to lock up.  I've tried 3 different GPSDO and they were all the same.  When it locks on, it is still not stable.
Why?  Because nature of GPSDO and Rb, time constant is set to 2 to 3 hours.  (still experimenting)  It takes few cycles of this to lock and stabilize.  In short, you'll be safe if you wait 24 to 48 hours.  Rb modules are relatively cheap.  I really don't see a need to baby it.


(Mr.) Taka Kamiya
I'm stuck in a wormhole....  Hello, worms!

On Friday, March 8, 2019, 4:00:34 AM EST, Forrest Christian (List Account) <lists@packetflux.com> wrote:  

Hopefully you'll all grace me with a few answers to a beginner
time-nut question or two.

I have a PRS-10 I've never used other than to power it on with a
recently-acquired heatsink and verify that it seems to operate
correctly and that the operational parameters don't seem out of
tolerance.  I would like to use this in the near future as a 10Mhz
reference for a TAPR TICC which I'd like to use to measure the jitter
performance of the PPS output of various consumer GPS receivers, the
goal being to end up with a jitter histogram.

So three interrelated questions:

  1. Assuming the PRS-10 has been off for a long time, how long should I
    plan on leaving this on for the 10Mhz to stabilize?  I see the
    longest warmup time on the spec-sheet is 7 minutes -  although this
    seems a lot shorter than I'd likely use in real life,  I'm also not
    sure if there's much benefit to an excessively longer warmup time
    (like days), would like opinions on this.

2)  Longer-term I'd like to use the 1PPS output from a Trimble
Thunderbolt to calibrate the PRS10 and adjust if necessary just to
trim out any aging drift on the PRS10.  Initially I thought I was
going to discipline the PRS10 on a continual basis with the
Thunderbolt using the PPS input on the PRS10, but I've recently
realized that leaving the PRS10 on permanently might not be the best
option (see Question 3).  So I'm looking for opinions on how to keep
the PRS10 calibrated/adjusted.  I.E. trim with the trimmer, adjust
using digital commands, etc.

  1. As implied in #2, I was originally planning on leaving the PRS10 on
    a continuous basis.  I've read a couple of things which imply that
    there is little benefit to doing so, and that every hour it's on
    consumes the lamp life.  Assuming I only need the highly stable PRS10
    source every few months for things like jitter measurements on 1PPS
    sources, is there any benefit to leaving the PRS10 on?

--

  • Forrest

time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

I am experimenting with exactly the same thing.  PRS-10 with Thunderbolt.  PRS-10 takes about a day to stabilize.  When physics lock happens, it has limited stability - which is mentioned in manual.  Before I calibrated, I waited for 2 days.  These movements are visible if you feed GPSDO into channel 1 of scope and set a trigger on it, and feed PRS-10 output to channel 2. For thunderbolt GPSDO, I am using a telecom unit, Nortel GPSTM modified with 1 pps output.  PRS-10 can take this input and lock, but it takes 1 to 2  hours to lock up.  I've tried 3 different GPSDO and they were all the same.  When it locks on, it is still not stable. Why?  Because nature of GPSDO and Rb, time constant is set to 2 to 3 hours.  (still experimenting)  It takes few cycles of this to lock and stabilize.  In short, you'll be safe if you wait 24 to 48 hours.  Rb modules are relatively cheap.  I really don't see a need to baby it. --------------------------------------- (Mr.) Taka Kamiya I'm stuck in a wormhole....  Hello, worms! On Friday, March 8, 2019, 4:00:34 AM EST, Forrest Christian (List Account) <lists@packetflux.com> wrote: Hopefully you'll all grace me with a few answers to a beginner time-nut question or two. I have a PRS-10 I've never used other than to power it on with a recently-acquired heatsink and verify that it seems to operate correctly and that the operational parameters don't seem out of tolerance.  I would like to use this in the near future as a 10Mhz reference for a TAPR TICC which I'd like to use to measure the jitter performance of the PPS output of various consumer GPS receivers, the goal being to end up with a jitter histogram. So three interrelated questions: 1) Assuming the PRS-10 has been off for a long time, how long should I plan on leaving this on for the 10Mhz to stabilize?  I see the longest warmup time on the spec-sheet is 7 minutes -  although this seems a lot shorter than I'd likely use in real life,  I'm also not sure if there's much benefit to an excessively longer warmup time (like days), would like opinions on this. 2)  Longer-term I'd like to use the 1PPS output from a Trimble Thunderbolt to calibrate the PRS10 and adjust if necessary just to trim out any aging drift on the PRS10.  Initially I thought I was going to discipline the PRS10 on a continual basis with the Thunderbolt using the PPS input on the PRS10, but I've recently realized that leaving the PRS10 on permanently might not be the best option (see Question 3).  So I'm looking for opinions on how to keep the PRS10 calibrated/adjusted.  I.E. trim with the trimmer, adjust using digital commands, etc. 3) As implied in #2, I was originally planning on leaving the PRS10 on a continuous basis.  I've read a couple of things which imply that there is little benefit to doing so, and that every hour it's on consumes the lamp life.  Assuming I only need the highly stable PRS10 source every few months for things like jitter measurements on 1PPS sources, is there any benefit to leaving the PRS10 on? -- - Forrest _______________________________________________ time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow the instructions there.
BK
Bob kb8tq
Fri, Mar 8, 2019 3:41 PM

Hi

On Mar 8, 2019, at 7:47 AM, Attila Kinali attila@kinali.ch wrote:

On Fri, 8 Mar 2019 01:18:45 -0700
"Forrest Christian (List Account)" lists@packetflux.com wrote:

  1. Assuming the PRS-10 has been off for a long time, how long should I
    plan on leaving this on for the 10Mhz to stabilize?  I see the
    longest warmup time on the spec-sheet is 7 minutes -  although this
    seems a lot shorter than I'd likely use in real life,  I'm also not
    sure if there's much benefit to an excessively longer warmup time
    (like days), would like opinions on this.

The warm-up time is from "I switched it on" to "there is a 10MHz signal
and it's 'stable'". I.e. at latest after 7 minutes you should be able
to use the ouptut of the PRS10. But keep in mind that the stable-after-warm-up
is not tha same as stable-for-a-time-nut. A Rb vapor cell standard has
a retrace (ie change of frequency after being switched on, for details see
John Vig's tutorial) of several weeks to a few months, depending on the
exact construction of the Rb cell, how long it has been in use and how
long it has been off.

  1. Longer-term I'd like to use the 1PPS output from a Trimble
    Thunderbolt to calibrate the PRS10 and adjust if necessary just to
    trim out any aging drift on the PRS10.  Initially I thought I was
    going to discipline the PRS10 on a continual basis with the
    Thunderbolt using the PPS input on the PRS10, but I've recently
    realized that leaving the PRS10 on permanently might not be the best
    option (see Question 3).  So I'm looking for opinions on how to keep
    the PRS10 calibrated/adjusted.  I.E. trim with the trimmer, adjust
    using digital commands, etc.

Switching it on and off several times a month would be worse in
terms of longevity  than having it running continuously. The vapor
cell itself is usually run between 60°C and 100°C and the Rb lamp
usually between 120°C and 170°C. As you can imagine, power cycling
gives a huge thermal strain on all the components. The inner construction
of the PRS10 is not optimized for power cycling but thermal insulation
of the sensitive components (see e.g. http://time.kinali.ch/Rb/PRS10/ )

So I would rather feed the PPS to the PRS10 directly and let itself
do the work. This will also give you an optimized control loop for
the performance of the PRS10 without you having to go through the
process of desiging and optimizing the loop.

  1. As implied in #2, I was originally planning on leaving the PRS10 on
    a continuous basis.  I've read a couple of things which imply that
    there is little benefit to doing so, and that every hour it's on
    consumes the lamp life.  Assuming I only need the highly stable PRS10
    source every few months for things like jitter measurements on 1PPS
    sources, is there any benefit to leaving the PRS10 on?

If you want to do just jitter measurement, then I wouldn't use a Rb standard
but a stable OCXO instead. At averaging times below 1-10s an OCXO will beat
any Rb standard, even one that has an OCXO like the PRS10 (admittedly, it's
not a high quality OCXO but it's one none-the-less). An OCXO has a shorter
warm-up time (usually 2-3 minutes until stable, usually less than a day
until time-nuts-stable, less than a month until nutty-time-nut-stable).

I personally, would use an OCXO as reference to the TICC (or TICCs?)
and use the PRS10 as an additional input to the TICC. This way you
can do a multi-way comparison between the GPSDOs and the Rb, removing
out the (in-)stability of the reference OCXO.

Putting some dimensions on this:

A “telecom” Rb should be in the 0.01 ppb at one second (ADEV) range. It should
improve by sqrt(tau) so at 100 seconds it will be at 0.001 ppb.  It should improve
some past that. Counting on 0.0001 ppb at 10,000 seconds …. not so much.

A cheap / small OCXO may be at the same 0.01 ppb at 1 second level. An OCXO
does not have a “predictable” improvement rate. You may have the same at 100
seconds as at 1 second.

A very good (but maybe cheap … who knows …) OCXO could get you to 0.001 ppb
at 1 second. You also could buy a few dozen to get one that actually does that. Testing
could be involved to sort them out.

Both the Rb and the OCXO have temperature stability specs on them. What those
numbers are may be unknown on a surplus item. The “best case” claimed temperature
stability on a telecom Rb is often better than the “best case” stability on your cheap
OCXO.

Indeed there are (rare) OCXO’s out there that get down to 0.0001 ppb  at 1 second
and hang in there out to 1,000 seconds. Finding one at all can be tough. Finding one
cheap … you would have to be very lucky. Working out that you do have one will involve
testing.

Bob

Beside those comments, you have not told us at what kind of performance
you are aiming at. There is a huge difference whether you need 1e-10 or
1e-13 and whether you need it at 1s, 1ks or even 100ks averaging times.

			Attila Kinali

--
It is upon moral qualities that a society is ultimately founded. All
the prosperity and technological sophistication in the world is of no
use without that foundation.
-- Miss Matheson, The Diamond Age, Neal Stephenson


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi > On Mar 8, 2019, at 7:47 AM, Attila Kinali <attila@kinali.ch> wrote: > > On Fri, 8 Mar 2019 01:18:45 -0700 > "Forrest Christian (List Account)" <lists@packetflux.com> wrote: > >> 1) Assuming the PRS-10 has been off for a long time, how long should I >> plan on leaving this on for the 10Mhz to stabilize? I see the >> longest warmup time on the spec-sheet is 7 minutes - although this >> seems a lot shorter than I'd likely use in real life, I'm also not >> sure if there's much benefit to an excessively longer warmup time >> (like days), would like opinions on this. > > The warm-up time is from "I switched it on" to "there is a 10MHz signal > and it's 'stable'". I.e. at latest after 7 minutes you should be able > to use the ouptut of the PRS10. But keep in mind that the stable-after-warm-up > is not tha same as stable-for-a-time-nut. A Rb vapor cell standard has > a retrace (ie change of frequency after being switched on, for details see > John Vig's tutorial) of several weeks to a few months, depending on the > exact construction of the Rb cell, how long it has been in use and how > long it has been off. > > >> 2) Longer-term I'd like to use the 1PPS output from a Trimble >> Thunderbolt to calibrate the PRS10 and adjust if necessary just to >> trim out any aging drift on the PRS10. Initially I thought I was >> going to discipline the PRS10 on a continual basis with the >> Thunderbolt using the PPS input on the PRS10, but I've recently >> realized that leaving the PRS10 on permanently might not be the best >> option (see Question 3). So I'm looking for opinions on how to keep >> the PRS10 calibrated/adjusted. I.E. trim with the trimmer, adjust >> using digital commands, etc. > > Switching it on and off several times a month would be worse in > terms of longevity than having it running continuously. The vapor > cell itself is usually run between 60°C and 100°C and the Rb lamp > usually between 120°C and 170°C. As you can imagine, power cycling > gives a huge thermal strain on all the components. The inner construction > of the PRS10 is not optimized for power cycling but thermal insulation > of the sensitive components (see e.g. http://time.kinali.ch/Rb/PRS10/ ) > > So I would rather feed the PPS to the PRS10 directly and let itself > do the work. This will also give you an optimized control loop for > the performance of the PRS10 without you having to go through the > process of desiging and optimizing the loop. > > >> 3) As implied in #2, I was originally planning on leaving the PRS10 on >> a continuous basis. I've read a couple of things which imply that >> there is little benefit to doing so, and that every hour it's on >> consumes the lamp life. Assuming I only need the highly stable PRS10 >> source every few months for things like jitter measurements on 1PPS >> sources, is there any benefit to leaving the PRS10 on? > > If you want to do just jitter measurement, then I wouldn't use a Rb standard > but a stable OCXO instead. At averaging times below 1-10s an OCXO will beat > any Rb standard, even one that has an OCXO like the PRS10 (admittedly, it's > not a high quality OCXO but it's one none-the-less). An OCXO has a shorter > warm-up time (usually 2-3 minutes until stable, usually less than a day > until time-nuts-stable, less than a month until nutty-time-nut-stable). > > I personally, would use an OCXO as reference to the TICC (or TICCs?) > and use the PRS10 as an additional input to the TICC. This way you > can do a multi-way comparison between the GPSDOs and the Rb, removing > out the (in-)stability of the reference OCXO. > Putting some dimensions on this: A “telecom” Rb should be in the 0.01 ppb at one second (ADEV) range. It should improve by sqrt(tau) so at 100 seconds it will be at 0.001 ppb. It should improve some past that. Counting on 0.0001 ppb at 10,000 seconds …. not so much. A cheap / small OCXO may be at the same 0.01 ppb at 1 second level. An OCXO does not have a “predictable” improvement rate. You may have the same at 100 seconds as at 1 second. A very good (but maybe cheap … who knows …) OCXO could get you to 0.001 ppb at 1 second. You also could buy a few dozen to get one that actually does that. Testing could be involved to sort them out. Both the Rb and the OCXO have temperature stability specs on them. What those numbers are may be unknown on a surplus item. The “best case” claimed temperature stability on a telecom Rb is often better than the “best case” stability on your cheap OCXO. Indeed there are (rare) OCXO’s out there that get down to 0.0001 ppb at 1 second and hang in there out to 1,000 seconds. Finding one at all can be tough. Finding one cheap … you would have to be very lucky. Working out that you *do* have one will involve testing. Bob > > Beside those comments, you have not told us at what kind of performance > you are aiming at. There is a huge difference whether you need 1e-10 or > 1e-13 and whether you need it at 1s, 1ks or even 100ks averaging times. > > > Attila Kinali > -- > It is upon moral qualities that a society is ultimately founded. All > the prosperity and technological sophistication in the world is of no > use without that foundation. > -- Miss Matheson, The Diamond Age, Neal Stephenson > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
AK
Attila Kinali
Fri, Mar 8, 2019 3:45 PM

On Fri, 8 Mar 2019 06:31:48 -0600
Dana Whitlow k8yumdoober@gmail.com wrote:

The point here is that there are apparently a number of warm up drift
mechanisms operating, some of which take days to sensibly settle down.

Longer. I know of one measurement, where the Rb had a kind of stable
drift until it suddenly switched to another slope quite suddenly
(within a few days) about half a year after power up.

The aging mechanisms of Rb vapor cell standards are many and not
all of them are well understood, much less controlled. Compared
to that, an OCXO has "only" thermal stability of the oven, strain
relaxation of the holder/crystal and deposition/removal of contaminants
on the crystal surface. Ok, there are a couple more, but these three
are the main contributors for most OCXO out there. While for the
Rb vapor cell standard I could name you half a dozen just like that
and I am far from being an expert on these.

For those interested, John Vig wrote a couple of papers on the aging
of OCXO in the 80s and 90s. The topic of Rb vapor cell aging is a lot
more messy and I don't know whether there is any good paper that reviews
the main contributors.

		Attila Kinali

--
It is upon moral qualities that a society is ultimately founded. All
the prosperity and technological sophistication in the world is of no
use without that foundation.
-- Miss Matheson, The Diamond Age, Neal Stephenson

On Fri, 8 Mar 2019 06:31:48 -0600 Dana Whitlow <k8yumdoober@gmail.com> wrote: > The point here is that there are apparently a number of warm up drift > mechanisms operating, some of which take days to sensibly settle down. Longer. I know of one measurement, where the Rb had a kind of stable drift until it suddenly switched to another slope quite suddenly (within a few days) about half a year after power up. The aging mechanisms of Rb vapor cell standards are many and not all of them are well understood, much less controlled. Compared to that, an OCXO has "only" thermal stability of the oven, strain relaxation of the holder/crystal and deposition/removal of contaminants on the crystal surface. Ok, there are a couple more, but these three are the main contributors for most OCXO out there. While for the Rb vapor cell standard I could name you half a dozen just like that and I am far from being an expert on these. For those interested, John Vig wrote a couple of papers on the aging of OCXO in the 80s and 90s. The topic of Rb vapor cell aging is a lot more messy and I don't know whether there is any good paper that reviews the main contributors. Attila Kinali -- It is upon moral qualities that a society is ultimately founded. All the prosperity and technological sophistication in the world is of no use without that foundation. -- Miss Matheson, The Diamond Age, Neal Stephenson
AK
Attila Kinali
Fri, Mar 8, 2019 3:56 PM

今日は

On Fri, 8 Mar 2019 14:09:20 +0000 (UTC)
Taka Kamiya via time-nuts time-nuts@lists.febo.com wrote:

Why?  Because nature of GPSDO and Rb, time constant is set to 2 to 3 hours. 
(still experimenting)  It takes few cycles of this to lock and stabilize.  In
short, you'll be safe if you wait 24 to 48 hours.  Rb modules are relatively
cheap.  I really don't see a need to baby it.

There are techniques to cut the "first lock time" down quite a bit.
Most of them involve either switching of all the long time constant
parts of the control loop or switching between different configurations
(ie different multiplicative constants or even different topologies).

The working principle of all is, that once you are reasonably close
in frequency/phase, your long-time-constant loop will not have a
large difference to work with and its internal state converges quickly
to something close of what it should be.

One very simple approach (which not always works) is to decompose your
PI controller into a P and I part, and disable on startup the I part.
Wait until the P part has gotten the error to something stable (not
neccesarily small) then switch on the I term, with its internal state
initialized to zero. A more sophisticated variant of this, uses an I
part with a smaller time constant first to reduce the residual error,
then estimates from its internal state the state of the real I part and
switches over to that.

Analysis of these dynamic loops can be a bit cumbersome and I recomend
to read a good textbook on this topic before attempting to do something
like this, as it's very easy to shot your foot.

			Attila Kinali

--
It is upon moral qualities that a society is ultimately founded. All
the prosperity and technological sophistication in the world is of no
use without that foundation.
-- Miss Matheson, The Diamond Age, Neal Stephenson

今日は On Fri, 8 Mar 2019 14:09:20 +0000 (UTC) Taka Kamiya via time-nuts <time-nuts@lists.febo.com> wrote: > Why?  Because nature of GPSDO and Rb, time constant is set to 2 to 3 hours.  > (still experimenting)  It takes few cycles of this to lock and stabilize.  In > short, you'll be safe if you wait 24 to 48 hours.  Rb modules are relatively > cheap.  I really don't see a need to baby it. There are techniques to cut the "first lock time" down quite a bit. Most of them involve either switching of all the long time constant parts of the control loop or switching between different configurations (ie different multiplicative constants or even different topologies). The working principle of all is, that once you are reasonably close in frequency/phase, your long-time-constant loop will not have a large difference to work with and its internal state converges quickly to something close of what it should be. One very simple approach (which not always works) is to decompose your PI controller into a P and I part, and disable on startup the I part. Wait until the P part has gotten the error to something stable (not neccesarily small) then switch on the I term, with its internal state initialized to zero. A more sophisticated variant of this, uses an I part with a smaller time constant first to reduce the residual error, then estimates from its internal state the state of the real I part and switches over to that. Analysis of these dynamic loops can be a bit cumbersome and I recomend to read a good textbook on this topic before attempting to do something like this, as it's very easy to shot your foot. Attila Kinali -- It is upon moral qualities that a society is ultimately founded. All the prosperity and technological sophistication in the world is of no use without that foundation. -- Miss Matheson, The Diamond Age, Neal Stephenson
MW
Michael Wouters
Sat, Mar 9, 2019 3:28 AM

We have operated  about 40 PRS10s over the past 20 years or so. These are
all run continuously in benign environments and monitored via GPS
time-transfer . Some have died after just over a year; others have operated
for more than 12 years. A new one generally takes a few months to burn in,
before its frequency drift stabilises to something like the specifications.
They do not behave very predictably, showing sudden changes in frequency
and so on. One exhibited frequent steps for a few years and then the
problem went away. Trying to predict what they would do at the level of
better than a few parts in 10^11 seems difficult.

Cheers
Michael

On Sat, 9 Mar 2019 at 3:03 am, Attila Kinali attila@kinali.ch wrote:

On Fri, 8 Mar 2019 06:31:48 -0600
Dana Whitlow k8yumdoober@gmail.com wrote:

The point here is that there are apparently a number of warm up drift
mechanisms operating, some of which take days to sensibly settle down.

Longer. I know of one measurement, where the Rb had a kind of stable
drift until it suddenly switched to another slope quite suddenly
(within a few days) about half a year after power up.

The aging mechanisms of Rb vapor cell standards are many and not
all of them are well understood, much less controlled. Compared
to that, an OCXO has "only" thermal stability of the oven, strain
relaxation of the holder/crystal and deposition/removal of contaminants
on the crystal surface. Ok, there are a couple more, but these three
are the main contributors for most OCXO out there. While for the
Rb vapor cell standard I could name you half a dozen just like that
and I am far from being an expert on these.

For those interested, John Vig wrote a couple of papers on the aging
of OCXO in the 80s and 90s. The topic of Rb vapor cell aging is a lot
more messy and I don't know whether there is any good paper that reviews
the main contributors.

                     Attila Kinali

--
It is upon moral qualities that a society is ultimately founded. All
the prosperity and technological sophistication in the world is of no
use without that foundation.
-- Miss Matheson, The Diamond Age, Neal Stephenson


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

We have operated about 40 PRS10s over the past 20 years or so. These are all run continuously in benign environments and monitored via GPS time-transfer . Some have died after just over a year; others have operated for more than 12 years. A new one generally takes a few months to burn in, before its frequency drift stabilises to something like the specifications. They do not behave very predictably, showing sudden changes in frequency and so on. One exhibited frequent steps for a few years and then the problem went away. Trying to predict what they would do at the level of better than a few parts in 10^11 seems difficult. Cheers Michael On Sat, 9 Mar 2019 at 3:03 am, Attila Kinali <attila@kinali.ch> wrote: > On Fri, 8 Mar 2019 06:31:48 -0600 > Dana Whitlow <k8yumdoober@gmail.com> wrote: > > > The point here is that there are apparently a number of warm up drift > > mechanisms operating, some of which take days to sensibly settle down. > > Longer. I know of one measurement, where the Rb had a kind of stable > drift until it suddenly switched to another slope quite suddenly > (within a few days) about half a year after power up. > > The aging mechanisms of Rb vapor cell standards are many and not > all of them are well understood, much less controlled. Compared > to that, an OCXO has "only" thermal stability of the oven, strain > relaxation of the holder/crystal and deposition/removal of contaminants > on the crystal surface. Ok, there are a couple more, but these three > are the main contributors for most OCXO out there. While for the > Rb vapor cell standard I could name you half a dozen just like that > and I am far from being an expert on these. > > For those interested, John Vig wrote a couple of papers on the aging > of OCXO in the 80s and 90s. The topic of Rb vapor cell aging is a lot > more messy and I don't know whether there is any good paper that reviews > the main contributors. > > Attila Kinali > -- > It is upon moral qualities that a society is ultimately founded. All > the prosperity and technological sophistication in the world is of no > use without that foundation. > -- Miss Matheson, The Diamond Age, Neal Stephenson > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >
FC
Forrest Christian (List Account)
Sat, Mar 9, 2019 7:04 AM

This seems like as good of a message to reply to as any.  Thanks to
everyone who has replied so far -  lots of good information, which is
right in the category of what I was looking for.

Let me clarify what I'm hoping to measure.  I apologize if some of
the terminology below is wrong, I'm still learning correct use of some
of the terms - I've tried to explain in a way where it is obvious what
I'm talking about even if the terms are slightly wrong.

The GPS receivers I'm hoping to measure are used in an application
where the 1PPS is used to synchronize radio transmissions.  Generally
one wants the pulse to be within a couple uS or so of the top of the
second.    Older radios cared less, newer ones want it to be more
precise.  Sorry about the soft spec, it unfortunately is what it is.
One note is that there is some expectation that there will be be
some fixed offset from the UTC second due to cable delays and the
like, so I'm less worried about a reasonably low fixed offset  and
more about the variable (jitter) offset.    I.E. 1uS +-100nS is
perfectly fine, 1uS +-1uS isn't so hot.

In evaluating a new set of GPS receivers for this application, I'm
hoping to gather Pulse-to-Pulse delay figures for every pulse the GPS
receiver outputs over a week or so.  I might end up repeating it in
various forms, including moving the GPS receiver into an
"environmental chamber" (aka modified chest freezer) to verify that
they still behave in low or high temperature - or if I can figure out
a reliable way to do so, to simulate various signal impairments such
as partial sky view.

My intent is to then process the raw data to get a feeling for the
statistical nature of the 1PPS out of that particular GPS receiver.
For instance, how many pulses were outside the +-ns spec for the
receiver.  What did the histogram of the in-spec pulses look like?
And so on. For some of the processing it's pretty obvious that I need
a clock source for the TICC which is going to be stable over the
entire period or else I'll end up measuring errors in the clock source
instead of the GPS receiver.    It seems to me that if I'm looking for
a histogram supposedly confined to +-100nS, I'm probably going to want
to be able to measure around +-1nS with confidence.  Much more than
that (0.1nS) seems a bit excessive, less than that (10nS) is too
coarse.

Intuition and past experience implies that if I want to measure +-1nS
with confidence, I really need a slightly more accurate clock source
than the measurement itself.  0.1ppb over the measurement period seems
reasonable, 0.2ppb is probably good enough.  I don't think I need
0.02ppb.

I replied in particular to Bob's message below since it seems to
confirm what I was thinking, even without me being verbose enough
about my application.  It also sounds like powering this on for a day
or so before I start (or getting started and just throwing the first
day's data away) might be reasonable to do and I don't need to let
this thing run for months ahead of time to get to the accuracy I need.
Of course, I'll do a bit of analysis on the data to try to verify
that the unit isn't drifting.

Out of curiosity, I'm also going to do some measurements on the couple
of standalone OCXO's in my collection to see if either would be
accurate enough for this particular measurement.  The best one in my
(admittedly limited) collection seems to be spec'd worse than I'd
need, although I'm not sure how much aging it's done.  None of them
have seen any meaningful poweron time just because I've been too busy
to play with them.  I guess that at some point I should also up my
game as far as good quality OCXO's since I do see how the right
well-aged low-drift OCXO would probably be fine for this measurement.

After reading everyone's response I'm still considering what to do as
far as long-term runtime on the PRS10.  I do know that I'm going to
need this off and on for the next few months so once I get it in a
permanent enclosure and get it powered up it will probably stay on
continuously at least until I am done with this round of experiments,
at which point it might get turned off with the hopes of extending
it's service life.

On Fri, Mar 8, 2019 at 8:06 AM Bob kb8tq kb8tq@n1k.org wrote:

Hi

Some of this gets into just what you are trying to do. Cycle to cycle jitter on the
PPS output of most GPS modules is in the “many ns” range. If sawtooth correction
is not being applied, a measurement that is good to an accuracy of 1 ns / 5 = 200 ps
would be plenty good enough. That is a 2x10^-10 (0.2 ppb) sort of accuracy.

If you are looking at longer term effects, you may need better accuracy / lower drift.
You can easily get past what a nice new 5071 can do heading down this road … For
now let’s assume the target above is correct.

The PRS-10 should be within 0.2 ppb within a few minutes after locking up. It should
hold this sort of accuracy “on the shelf / power off” pretty much forever and ever. In fact,
it’s a pretty good bet that it will hold 10X this accuracy for quite a long time. You probably
should warm it up for a day or so if you are after 0.02 ppb.

Calibration wise, the “easy” way to do it depends a lot on what you have to do the measurement.
If you have a TICC (sounds like you do) - use it. You can also use a scope or a more
conventional counter.

If you are going after 0.02 ppb, you would want to set the unit to 0.002 ppb. That gives you
90% of the budget for things like temperature drift and aging. Indeed you might want to go
to a tighter number …. this is Time Nuts after all.

0.002 ppb is 2 ns over 1,000 seconds. It’s 200 ns over 100,000 seconds. (Amazing how that
works :) ). Is your TBolt PPS good to 2 ns? Probably not. Is it good to 200 ns? most certainly.
Since 100,000 seconds is just over a day, that makes for a fairly easy adjustment process.
Look at it once a day and tweak it. After a while it will not need tweaks any more.

The advantage of a “long term” adjustment like this is that it also takes in the temperature
swings in your lab over a one day period. At the 0.002 ppb level, they will be the dominant
part of what you see. Will you get to the 0.002 ppb level? It depends a lot on just how stable
your particular unit is and how drafty your lab is.

Running any sort of electronics gear 24 hours a day is a risk. It may wear out, it may cause
a fire, it may be fine … who knows. It also pulls power and you pay the electric company for that.
It’s generally cheaper / safer to turn stuff off if it is only going to be used rarely.

Lots of fun !!!

Bob

On Mar 8, 2019, at 3:18 AM, Forrest Christian (List Account) lists@packetflux.com wrote:

Hopefully you'll all grace me with a few answers to a beginner
time-nut question or two.

I have a PRS-10 I've never used other than to power it on with a
recently-acquired heatsink and verify that it seems to operate
correctly and that the operational parameters don't seem out of
tolerance.  I would like to use this in the near future as a 10Mhz
reference for a TAPR TICC which I'd like to use to measure the jitter
performance of the PPS output of various consumer GPS receivers, the
goal being to end up with a jitter histogram.

So three interrelated questions:

  1. Assuming the PRS-10 has been off for a long time, how long should I
    plan on leaving this on for the 10Mhz to stabilize?  I see the
    longest warmup time on the spec-sheet is 7 minutes -  although this
    seems a lot shorter than I'd likely use in real life,  I'm also not
    sure if there's much benefit to an excessively longer warmup time
    (like days), would like opinions on this.

  2. Longer-term I'd like to use the 1PPS output from a Trimble
    Thunderbolt to calibrate the PRS10 and adjust if necessary just to
    trim out any aging drift on the PRS10.  Initially I thought I was
    going to discipline the PRS10 on a continual basis with the
    Thunderbolt using the PPS input on the PRS10, but I've recently
    realized that leaving the PRS10 on permanently might not be the best
    option (see Question 3).  So I'm looking for opinions on how to keep
    the PRS10 calibrated/adjusted.  I.E. trim with the trimmer, adjust
    using digital commands, etc.

  3. As implied in #2, I was originally planning on leaving the PRS10 on
    a continuous basis.  I've read a couple of things which imply that
    there is little benefit to doing so, and that every hour it's on
    consumes the lamp life.  Assuming I only need the highly stable PRS10
    source every few months for things like jitter measurements on 1PPS
    sources, is there any benefit to leaving the PRS10 on?

--

  • Forrest

time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

--

  • Forrest
This seems like as good of a message to reply to as any. Thanks to everyone who has replied so far - lots of good information, which is right in the category of what I was looking for. Let me clarify what I'm hoping to measure. I apologize if some of the terminology below is wrong, I'm still learning correct use of some of the terms - I've tried to explain in a way where it is obvious what I'm talking about even if the terms are slightly wrong. The GPS receivers I'm hoping to measure are used in an application where the 1PPS is used to synchronize radio transmissions. Generally one wants the pulse to be within a couple uS or so of the top of the second. Older radios cared less, newer ones want it to be more precise. Sorry about the soft spec, it unfortunately is what it is. One note is that there is some expectation that there will be be some fixed offset from the UTC second due to cable delays and the like, so I'm less worried about a reasonably low fixed offset and more about the variable (jitter) offset. I.E. 1uS +-100nS is perfectly fine, 1uS +-1uS isn't so hot. In evaluating a new set of GPS receivers for this application, I'm hoping to gather Pulse-to-Pulse delay figures for every pulse the GPS receiver outputs over a week or so. I might end up repeating it in various forms, including moving the GPS receiver into an "environmental chamber" (aka modified chest freezer) to verify that they still behave in low or high temperature - or if I can figure out a reliable way to do so, to simulate various signal impairments such as partial sky view. My intent is to then process the raw data to get a feeling for the statistical nature of the 1PPS out of that particular GPS receiver. For instance, how many pulses were outside the +-ns spec for the receiver. What did the histogram of the in-spec pulses look like? And so on. For some of the processing it's pretty obvious that I need a clock source for the TICC which is going to be stable over the entire period or else I'll end up measuring errors in the clock source instead of the GPS receiver. It seems to me that if I'm looking for a histogram supposedly confined to +-100nS, I'm probably going to want to be able to measure around +-1nS with confidence. Much more than that (0.1nS) seems a bit excessive, less than that (10nS) is too coarse. Intuition and past experience implies that if I want to measure +-1nS with confidence, I really need a slightly more accurate clock source than the measurement itself. 0.1ppb over the measurement period seems reasonable, 0.2ppb is probably good enough. I don't think I need 0.02ppb. I replied in particular to Bob's message below since it seems to confirm what I was thinking, even without me being verbose enough about my application. It also sounds like powering this on for a day or so before I start (or getting started and just throwing the first day's data away) might be reasonable to do and I don't need to let this thing run for months ahead of time to get to the accuracy I need. Of course, I'll do a bit of analysis on the data to try to verify that the unit isn't drifting. Out of curiosity, I'm also going to do some measurements on the couple of standalone OCXO's in my collection to see if either would be accurate enough for this particular measurement. The best one in my (admittedly limited) collection seems to be spec'd worse than I'd need, although I'm not sure how much aging it's done. None of them have seen any meaningful poweron time just because I've been too busy to play with them. I guess that at some point I should also up my game as far as good quality OCXO's since I do see how the right well-aged low-drift OCXO would probably be fine for this measurement. After reading everyone's response I'm still considering what to do as far as long-term runtime on the PRS10. I do know that I'm going to need this off and on for the next few months so once I get it in a permanent enclosure and get it powered up it will probably stay on continuously at least until I am done with this round of experiments, at which point it might get turned off with the hopes of extending it's service life. On Fri, Mar 8, 2019 at 8:06 AM Bob kb8tq <kb8tq@n1k.org> wrote: > > Hi > > Some of this gets into just what you are trying to do. Cycle to cycle jitter on the > PPS output of most GPS modules is in the “many ns” range. If sawtooth correction > is not being applied, a measurement that is good to an accuracy of 1 ns / 5 = 200 ps > would be plenty good enough. That is a 2x10^-10 (0.2 ppb) sort of accuracy. > > If you are looking at longer term effects, you may need better accuracy / lower drift. > You can easily get past what a nice new 5071 can do heading down this road … For > now let’s assume the target above is correct. > > The PRS-10 should be within 0.2 ppb within a few minutes after locking up. It should > hold this sort of accuracy “on the shelf / power off” pretty much forever and ever. In fact, > it’s a pretty good bet that it will hold 10X this accuracy for quite a long time. You probably > should warm it up for a day or so if you are after 0.02 ppb. > > Calibration wise, the “easy” way to do it depends a lot on what you have to do the measurement. > If you have a TICC (sounds like you do) - use it. You can also use a scope or a more > conventional counter. > > If you are going after 0.02 ppb, you would want to set the unit to 0.002 ppb. That gives you > 90% of the budget for things like temperature drift and aging. Indeed you might want to go > to a tighter number …. this is Time Nuts after all. > > 0.002 ppb is 2 ns over 1,000 seconds. It’s 200 ns over 100,000 seconds. (Amazing how that > works :) ). Is your TBolt PPS good to 2 ns? Probably not. Is it good to 200 ns? most certainly. > Since 100,000 seconds is just over a day, that makes for a fairly easy adjustment process. > Look at it once a day and tweak it. After a while it will not need tweaks any more. > > The advantage of a “long term” adjustment like this is that it also takes in the temperature > swings in your lab over a one day period. At the 0.002 ppb level, they will be the dominant > part of what you see. Will you get to the 0.002 ppb level? It depends a lot on just how stable > your particular unit is and how drafty your lab is. > > Running any sort of electronics gear 24 hours a day is a risk. It may wear out, it may cause > a fire, it may be fine … who knows. It also pulls power and you pay the electric company for that. > It’s generally cheaper / safer to turn stuff off if it is only going to be used rarely. > > Lots of fun !!! > > Bob > > > On Mar 8, 2019, at 3:18 AM, Forrest Christian (List Account) <lists@packetflux.com> wrote: > > > > Hopefully you'll all grace me with a few answers to a beginner > > time-nut question or two. > > > > I have a PRS-10 I've never used other than to power it on with a > > recently-acquired heatsink and verify that it seems to operate > > correctly and that the operational parameters don't seem out of > > tolerance. I would like to use this in the near future as a 10Mhz > > reference for a TAPR TICC which I'd like to use to measure the jitter > > performance of the PPS output of various consumer GPS receivers, the > > goal being to end up with a jitter histogram. > > > > So three interrelated questions: > > > > 1) Assuming the PRS-10 has been off for a long time, how long should I > > plan on leaving this on for the 10Mhz to stabilize? I see the > > longest warmup time on the spec-sheet is 7 minutes - although this > > seems a lot shorter than I'd likely use in real life, I'm also not > > sure if there's much benefit to an excessively longer warmup time > > (like days), would like opinions on this. > > > > 2) Longer-term I'd like to use the 1PPS output from a Trimble > > Thunderbolt to calibrate the PRS10 and adjust if necessary just to > > trim out any aging drift on the PRS10. Initially I thought I was > > going to discipline the PRS10 on a continual basis with the > > Thunderbolt using the PPS input on the PRS10, but I've recently > > realized that leaving the PRS10 on permanently might not be the best > > option (see Question 3). So I'm looking for opinions on how to keep > > the PRS10 calibrated/adjusted. I.E. trim with the trimmer, adjust > > using digital commands, etc. > > > > 3) As implied in #2, I was originally planning on leaving the PRS10 on > > a continuous basis. I've read a couple of things which imply that > > there is little benefit to doing so, and that every hour it's on > > consumes the lamp life. Assuming I only need the highly stable PRS10 > > source every few months for things like jitter measurements on 1PPS > > sources, is there any benefit to leaving the PRS10 on? > > > > -- > > - Forrest > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@lists.febo.com > > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. -- - Forrest
BK
Bob kb8tq
Sat, Mar 9, 2019 1:15 PM

Hi

Ok, if you are concerned about 1 us and not to concerned about 100 ns, then there’s not
a lot to set up. A standard that is good to 10 ns / s will do just fine. That’s 10 ppb and a
“typical” OCXO in a counter will do just fine. Leave it on for a few days and set it to frequency.

There are maybe a few hundred GPS module manufacturers out there. Going through all the
in’s and out’s to fiddle with each one is a major task. There are timing grade GPS modules
out there They are “good” in the 10’s of ns without any real crazy stuff. They are good in the
single digit ns with some effort.

Unless this is a very large volume product, just use one of the well known parts and move
on. The uBlox M8T is a modern part that will easily meet your needs. Furuno and others make
parts that are at least as good. Cost wise, you are <$50 for small volume and much less in large
volume.

I would not bother with a part that has not been designed to do timing. They can have bugs
in them that only manifest at the worst moments. (like two months later in the field). At the
very least, start with a known good part to get your setup going before you mess with any
thing else.

========

These devices (at least the timing ones) all have delay adjustments on them. You can zero
out any cable delays. The same is true of a TBolt. You can set it’s PPS pulse to be early
or late relative to GPS time (= no leap seconds, UTC has leap seconds).

This adjustment also lets you come up with a much easier way to test your modules.

Take the TBolt and set it to be 1 us early.

Put the TBolt into channel A and the device you are testing into channel B on your counter.

Measure the time between the two channels.

You now don’t have to worry about the standard on the counter, it’s only running for one us.

The TBolt is spec’d to be < +/- 100 ns of GPS all the time. They normally wobble around
by < 10 ns.

Lots of fun !!!

Bob

On Mar 9, 2019, at 2:04 AM, Forrest Christian (List Account) lists@packetflux.com wrote:

This seems like as good of a message to reply to as any.  Thanks to
everyone who has replied so far -  lots of good information, which is
right in the category of what I was looking for.

Let me clarify what I'm hoping to measure.  I apologize if some of
the terminology below is wrong, I'm still learning correct use of some
of the terms - I've tried to explain in a way where it is obvious what
I'm talking about even if the terms are slightly wrong.

The GPS receivers I'm hoping to measure are used in an application
where the 1PPS is used to synchronize radio transmissions.  Generally
one wants the pulse to be within a couple uS or so of the top of the
second.    Older radios cared less, newer ones want it to be more
precise.  Sorry about the soft spec, it unfortunately is what it is.
One note is that there is some expectation that there will be be
some fixed offset from the UTC second due to cable delays and the
like, so I'm less worried about a reasonably low fixed offset  and
more about the variable (jitter) offset.    I.E. 1uS +-100nS is
perfectly fine, 1uS +-1uS isn't so hot.

In evaluating a new set of GPS receivers for this application, I'm
hoping to gather Pulse-to-Pulse delay figures for every pulse the GPS
receiver outputs over a week or so.  I might end up repeating it in
various forms, including moving the GPS receiver into an
"environmental chamber" (aka modified chest freezer) to verify that
they still behave in low or high temperature - or if I can figure out
a reliable way to do so, to simulate various signal impairments such
as partial sky view.

My intent is to then process the raw data to get a feeling for the
statistical nature of the 1PPS out of that particular GPS receiver.
For instance, how many pulses were outside the +-ns spec for the
receiver.  What did the histogram of the in-spec pulses look like?
And so on. For some of the processing it's pretty obvious that I need
a clock source for the TICC which is going to be stable over the
entire period or else I'll end up measuring errors in the clock source
instead of the GPS receiver.    It seems to me that if I'm looking for
a histogram supposedly confined to +-100nS, I'm probably going to want
to be able to measure around +-1nS with confidence.  Much more than
that (0.1nS) seems a bit excessive, less than that (10nS) is too
coarse.

Intuition and past experience implies that if I want to measure +-1nS
with confidence, I really need a slightly more accurate clock source
than the measurement itself.  0.1ppb over the measurement period seems
reasonable, 0.2ppb is probably good enough.  I don't think I need
0.02ppb.

I replied in particular to Bob's message below since it seems to
confirm what I was thinking, even without me being verbose enough
about my application.  It also sounds like powering this on for a day
or so before I start (or getting started and just throwing the first
day's data away) might be reasonable to do and I don't need to let
this thing run for months ahead of time to get to the accuracy I need.
Of course, I'll do a bit of analysis on the data to try to verify
that the unit isn't drifting.

Out of curiosity, I'm also going to do some measurements on the couple
of standalone OCXO's in my collection to see if either would be
accurate enough for this particular measurement.  The best one in my
(admittedly limited) collection seems to be spec'd worse than I'd
need, although I'm not sure how much aging it's done.  None of them
have seen any meaningful poweron time just because I've been too busy
to play with them.  I guess that at some point I should also up my
game as far as good quality OCXO's since I do see how the right
well-aged low-drift OCXO would probably be fine for this measurement.

After reading everyone's response I'm still considering what to do as
far as long-term runtime on the PRS10.  I do know that I'm going to
need this off and on for the next few months so once I get it in a
permanent enclosure and get it powered up it will probably stay on
continuously at least until I am done with this round of experiments,
at which point it might get turned off with the hopes of extending
it's service life.

On Fri, Mar 8, 2019 at 8:06 AM Bob kb8tq kb8tq@n1k.org wrote:

Hi

Some of this gets into just what you are trying to do. Cycle to cycle jitter on the
PPS output of most GPS modules is in the “many ns” range. If sawtooth correction
is not being applied, a measurement that is good to an accuracy of 1 ns / 5 = 200 ps
would be plenty good enough. That is a 2x10^-10 (0.2 ppb) sort of accuracy.

If you are looking at longer term effects, you may need better accuracy / lower drift.
You can easily get past what a nice new 5071 can do heading down this road … For
now let’s assume the target above is correct.

The PRS-10 should be within 0.2 ppb within a few minutes after locking up. It should
hold this sort of accuracy “on the shelf / power off” pretty much forever and ever. In fact,
it’s a pretty good bet that it will hold 10X this accuracy for quite a long time. You probably
should warm it up for a day or so if you are after 0.02 ppb.

Calibration wise, the “easy” way to do it depends a lot on what you have to do the measurement.
If you have a TICC (sounds like you do) - use it. You can also use a scope or a more
conventional counter.

If you are going after 0.02 ppb, you would want to set the unit to 0.002 ppb. That gives you
90% of the budget for things like temperature drift and aging. Indeed you might want to go
to a tighter number …. this is Time Nuts after all.

0.002 ppb is 2 ns over 1,000 seconds. It’s 200 ns over 100,000 seconds. (Amazing how that
works :) ). Is your TBolt PPS good to 2 ns? Probably not. Is it good to 200 ns? most certainly.
Since 100,000 seconds is just over a day, that makes for a fairly easy adjustment process.
Look at it once a day and tweak it. After a while it will not need tweaks any more.

The advantage of a “long term” adjustment like this is that it also takes in the temperature
swings in your lab over a one day period. At the 0.002 ppb level, they will be the dominant
part of what you see. Will you get to the 0.002 ppb level? It depends a lot on just how stable
your particular unit is and how drafty your lab is.

Running any sort of electronics gear 24 hours a day is a risk. It may wear out, it may cause
a fire, it may be fine … who knows. It also pulls power and you pay the electric company for that.
It’s generally cheaper / safer to turn stuff off if it is only going to be used rarely.

Lots of fun !!!

Bob

On Mar 8, 2019, at 3:18 AM, Forrest Christian (List Account) lists@packetflux.com wrote:

Hopefully you'll all grace me with a few answers to a beginner
time-nut question or two.

I have a PRS-10 I've never used other than to power it on with a
recently-acquired heatsink and verify that it seems to operate
correctly and that the operational parameters don't seem out of
tolerance.  I would like to use this in the near future as a 10Mhz
reference for a TAPR TICC which I'd like to use to measure the jitter
performance of the PPS output of various consumer GPS receivers, the
goal being to end up with a jitter histogram.

So three interrelated questions:

  1. Assuming the PRS-10 has been off for a long time, how long should I
    plan on leaving this on for the 10Mhz to stabilize?  I see the
    longest warmup time on the spec-sheet is 7 minutes -  although this
    seems a lot shorter than I'd likely use in real life,  I'm also not
    sure if there's much benefit to an excessively longer warmup time
    (like days), would like opinions on this.

  2. Longer-term I'd like to use the 1PPS output from a Trimble
    Thunderbolt to calibrate the PRS10 and adjust if necessary just to
    trim out any aging drift on the PRS10.  Initially I thought I was
    going to discipline the PRS10 on a continual basis with the
    Thunderbolt using the PPS input on the PRS10, but I've recently
    realized that leaving the PRS10 on permanently might not be the best
    option (see Question 3).  So I'm looking for opinions on how to keep
    the PRS10 calibrated/adjusted.  I.E. trim with the trimmer, adjust
    using digital commands, etc.

  3. As implied in #2, I was originally planning on leaving the PRS10 on
    a continuous basis.  I've read a couple of things which imply that
    there is little benefit to doing so, and that every hour it's on
    consumes the lamp life.  Assuming I only need the highly stable PRS10
    source every few months for things like jitter measurements on 1PPS
    sources, is there any benefit to leaving the PRS10 on?

--

  • Forrest

time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

--

  • Forrest

time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi Ok, if you are concerned about 1 us and not to concerned about 100 ns, then there’s not a lot to set up. A standard that is good to 10 ns / s will do just fine. That’s 10 ppb and a “typical” OCXO in a counter will do just fine. Leave it on for a few days and set it to frequency. There are maybe a few hundred GPS module manufacturers out there. Going through all the in’s and out’s to fiddle with each one is a major task. There *are* timing grade GPS modules out there They are “good” in the 10’s of ns without any real crazy stuff. They are good in the single digit ns with some effort. Unless this is a very large volume product, just use one of the well known parts and move on. The uBlox M8T is a modern part that will easily meet your needs. Furuno and others make parts that are at least as good. Cost wise, you are <$50 for small volume and much less in large volume. I would not bother with a part that has not been designed to do timing. They can have bugs in them that only manifest at the worst moments. (like two months later in the field). At the very least, start with a known good part to get your setup going before you mess with any thing else. ======== These devices (at least the timing ones) all have delay adjustments on them. You can zero out any cable delays. The same is true of a TBolt. You can set it’s PPS pulse to be early or late relative to GPS time (= no leap seconds, UTC has leap seconds). This adjustment also lets you come up with a *much* easier way to test your modules. Take the TBolt and set it to be 1 us early. Put the TBolt into channel A and the device you are testing into channel B on your counter. Measure the time between the two channels. You now don’t have to worry about the standard on the counter, it’s only running for one us. The TBolt is spec’d to be < +/- 100 ns of GPS all the time. They normally wobble around by < 10 ns. Lots of fun !!! Bob > On Mar 9, 2019, at 2:04 AM, Forrest Christian (List Account) <lists@packetflux.com> wrote: > > This seems like as good of a message to reply to as any. Thanks to > everyone who has replied so far - lots of good information, which is > right in the category of what I was looking for. > > Let me clarify what I'm hoping to measure. I apologize if some of > the terminology below is wrong, I'm still learning correct use of some > of the terms - I've tried to explain in a way where it is obvious what > I'm talking about even if the terms are slightly wrong. > > The GPS receivers I'm hoping to measure are used in an application > where the 1PPS is used to synchronize radio transmissions. Generally > one wants the pulse to be within a couple uS or so of the top of the > second. Older radios cared less, newer ones want it to be more > precise. Sorry about the soft spec, it unfortunately is what it is. > One note is that there is some expectation that there will be be > some fixed offset from the UTC second due to cable delays and the > like, so I'm less worried about a reasonably low fixed offset and > more about the variable (jitter) offset. I.E. 1uS +-100nS is > perfectly fine, 1uS +-1uS isn't so hot. > > In evaluating a new set of GPS receivers for this application, I'm > hoping to gather Pulse-to-Pulse delay figures for every pulse the GPS > receiver outputs over a week or so. I might end up repeating it in > various forms, including moving the GPS receiver into an > "environmental chamber" (aka modified chest freezer) to verify that > they still behave in low or high temperature - or if I can figure out > a reliable way to do so, to simulate various signal impairments such > as partial sky view. > > My intent is to then process the raw data to get a feeling for the > statistical nature of the 1PPS out of that particular GPS receiver. > For instance, how many pulses were outside the +-ns spec for the > receiver. What did the histogram of the in-spec pulses look like? > And so on. For some of the processing it's pretty obvious that I need > a clock source for the TICC which is going to be stable over the > entire period or else I'll end up measuring errors in the clock source > instead of the GPS receiver. It seems to me that if I'm looking for > a histogram supposedly confined to +-100nS, I'm probably going to want > to be able to measure around +-1nS with confidence. Much more than > that (0.1nS) seems a bit excessive, less than that (10nS) is too > coarse. > > Intuition and past experience implies that if I want to measure +-1nS > with confidence, I really need a slightly more accurate clock source > than the measurement itself. 0.1ppb over the measurement period seems > reasonable, 0.2ppb is probably good enough. I don't think I need > 0.02ppb. > > I replied in particular to Bob's message below since it seems to > confirm what I was thinking, even without me being verbose enough > about my application. It also sounds like powering this on for a day > or so before I start (or getting started and just throwing the first > day's data away) might be reasonable to do and I don't need to let > this thing run for months ahead of time to get to the accuracy I need. > Of course, I'll do a bit of analysis on the data to try to verify > that the unit isn't drifting. > > Out of curiosity, I'm also going to do some measurements on the couple > of standalone OCXO's in my collection to see if either would be > accurate enough for this particular measurement. The best one in my > (admittedly limited) collection seems to be spec'd worse than I'd > need, although I'm not sure how much aging it's done. None of them > have seen any meaningful poweron time just because I've been too busy > to play with them. I guess that at some point I should also up my > game as far as good quality OCXO's since I do see how the right > well-aged low-drift OCXO would probably be fine for this measurement. > > After reading everyone's response I'm still considering what to do as > far as long-term runtime on the PRS10. I do know that I'm going to > need this off and on for the next few months so once I get it in a > permanent enclosure and get it powered up it will probably stay on > continuously at least until I am done with this round of experiments, > at which point it might get turned off with the hopes of extending > it's service life. > > On Fri, Mar 8, 2019 at 8:06 AM Bob kb8tq <kb8tq@n1k.org> wrote: >> >> Hi >> >> Some of this gets into just what you are trying to do. Cycle to cycle jitter on the >> PPS output of most GPS modules is in the “many ns” range. If sawtooth correction >> is not being applied, a measurement that is good to an accuracy of 1 ns / 5 = 200 ps >> would be plenty good enough. That is a 2x10^-10 (0.2 ppb) sort of accuracy. >> >> If you are looking at longer term effects, you may need better accuracy / lower drift. >> You can easily get past what a nice new 5071 can do heading down this road … For >> now let’s assume the target above is correct. >> >> The PRS-10 should be within 0.2 ppb within a few minutes after locking up. It should >> hold this sort of accuracy “on the shelf / power off” pretty much forever and ever. In fact, >> it’s a pretty good bet that it will hold 10X this accuracy for quite a long time. You probably >> should warm it up for a day or so if you are after 0.02 ppb. >> >> Calibration wise, the “easy” way to do it depends a lot on what you have to do the measurement. >> If you have a TICC (sounds like you do) - use it. You can also use a scope or a more >> conventional counter. >> >> If you are going after 0.02 ppb, you would want to set the unit to 0.002 ppb. That gives you >> 90% of the budget for things like temperature drift and aging. Indeed you might want to go >> to a tighter number …. this is Time Nuts after all. >> >> 0.002 ppb is 2 ns over 1,000 seconds. It’s 200 ns over 100,000 seconds. (Amazing how that >> works :) ). Is your TBolt PPS good to 2 ns? Probably not. Is it good to 200 ns? most certainly. >> Since 100,000 seconds is just over a day, that makes for a fairly easy adjustment process. >> Look at it once a day and tweak it. After a while it will not need tweaks any more. >> >> The advantage of a “long term” adjustment like this is that it also takes in the temperature >> swings in your lab over a one day period. At the 0.002 ppb level, they will be the dominant >> part of what you see. Will you get to the 0.002 ppb level? It depends a lot on just how stable >> your particular unit is and how drafty your lab is. >> >> Running any sort of electronics gear 24 hours a day is a risk. It may wear out, it may cause >> a fire, it may be fine … who knows. It also pulls power and you pay the electric company for that. >> It’s generally cheaper / safer to turn stuff off if it is only going to be used rarely. >> >> Lots of fun !!! >> >> Bob >> >>> On Mar 8, 2019, at 3:18 AM, Forrest Christian (List Account) <lists@packetflux.com> wrote: >>> >>> Hopefully you'll all grace me with a few answers to a beginner >>> time-nut question or two. >>> >>> I have a PRS-10 I've never used other than to power it on with a >>> recently-acquired heatsink and verify that it seems to operate >>> correctly and that the operational parameters don't seem out of >>> tolerance. I would like to use this in the near future as a 10Mhz >>> reference for a TAPR TICC which I'd like to use to measure the jitter >>> performance of the PPS output of various consumer GPS receivers, the >>> goal being to end up with a jitter histogram. >>> >>> So three interrelated questions: >>> >>> 1) Assuming the PRS-10 has been off for a long time, how long should I >>> plan on leaving this on for the 10Mhz to stabilize? I see the >>> longest warmup time on the spec-sheet is 7 minutes - although this >>> seems a lot shorter than I'd likely use in real life, I'm also not >>> sure if there's much benefit to an excessively longer warmup time >>> (like days), would like opinions on this. >>> >>> 2) Longer-term I'd like to use the 1PPS output from a Trimble >>> Thunderbolt to calibrate the PRS10 and adjust if necessary just to >>> trim out any aging drift on the PRS10. Initially I thought I was >>> going to discipline the PRS10 on a continual basis with the >>> Thunderbolt using the PPS input on the PRS10, but I've recently >>> realized that leaving the PRS10 on permanently might not be the best >>> option (see Question 3). So I'm looking for opinions on how to keep >>> the PRS10 calibrated/adjusted. I.E. trim with the trimmer, adjust >>> using digital commands, etc. >>> >>> 3) As implied in #2, I was originally planning on leaving the PRS10 on >>> a continuous basis. I've read a couple of things which imply that >>> there is little benefit to doing so, and that every hour it's on >>> consumes the lamp life. Assuming I only need the highly stable PRS10 >>> source every few months for things like jitter measurements on 1PPS >>> sources, is there any benefit to leaving the PRS10 on? >>> >>> -- >>> - Forrest >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@lists.febo.com >>> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>> and follow the instructions there. >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. > > > > -- > - Forrest > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.