Hi All.
I have an SR620 that is misbehaving. I am working my way through the manual
and schematics, but one consistent behavior has arisen: At a period of
around 4-6s when measuring frequency, I get a "wrong number" that is
consistent in scale across 10MHz and 1kHz (the REF frequency).
I.e. when I measure a 10MHz signal, occasionally the display shows
9946187.4... Hz and when measuring REF (1kHz) the display occasionally
shows 994.618757 Hz. This is independent of termination, and the A & B
front-ends have passed their tests.
Has anyone encountered the same issue, or have a suggestion as to where to
look for the root cause?
[image: 20221104_SR620_SN356_Frequency_error-1-small.jpg]
[image: 20221104_SR620_SN356_Frequency_error-2-small.jpg]
[image: 20221104_SR620_SN356_Frequency_error-3-small.jpg]
--Andrew
Andrew E. Kalman, Ph.D.
Hi Andrew,
On 11/4/22 19:44, Andrew Kalman via time-nuts wrote:
Hi All.
I have an SR620 that is misbehaving. I am working my way through the manual
and schematics, but one consistent behavior has arisen: At a period of
around 4-6s when measuring frequency, I get a "wrong number" that is
consistent in scale across 10MHz and 1kHz (the REF frequency).
I.e. when I measure a 10MHz signal, occasionally the display shows
9946187.4... Hz and when measuring REF (1kHz) the display occasionally
shows 994.618757 Hz. This is independent of termination, and the A & B
front-ends have passed their tests.
Has anyone encountered the same issue, or have a suggestion as to where to
look for the root cause?
[image: 20221104_SR620_SN356_Frequency_error-1-small.jpg]
[image: 20221104_SR620_SN356_Frequency_error-2-small.jpg]
[image: 20221104_SR620_SN356_Frequency_error-3-small.jpg]
Since f = cycles / time, these two latest is actually about the same time.
time = cycles / f and cycles is nominal frequency times nominal
time-base, so
t1 = 1E5 / 9946187.4882 = .0100541036 s
t2 = 1E1 / 994.618757 = .01005410357 s
So, both are really taking about the same time, just above the 10 ms the
time-base is set for, 54.1 us or so.
I would guess that there is an issue with the stop-counter side, so it
somehow keeps running. The difference you see is probably in the
different start-time-stamps.
The time as used in above formula is really stop-time minus start-time
to get the elapsed time, and the time-base triggers how far after the
start-time that an attempt at a stop-time is taken. So, it's not
unreasonable that we are just above 10 ms, it's the intended behavior,
but the actual values is consistent with some issue to it.
I hope this gives you a good clue of what part of the good old SR620
troubling you.
Sometimes I find that tweaking the trigger-points avoid these issues.
That is part of the working that turns into part of a skill. However,
when your trigger is in auto for normal good signals, you should not
experience this. If messing with your trigger-point help you, it could
be that you need to work on the trigger side of things to get stable
operation as default properly. The 1 kHz output on loopback should
always operate safely in auto-mode, yet you illustrate it's failure. I
seem to recall calibration details about it.
Cheers,
Magnus
Magnus: Thank you, I will go down that path. Your explanation makes sense,
and I figured it was probably something with the start and stop times, but
didn't see the math behind it.
FWIW, on this unit:
I designed the SRS SR560 (everything but the front-end, that was founder
John Willison's design) as my first task at SRS. My buddies Scott and Adam
did the PCB and enclosure of the SR620 (and many other models),
respectively, and then we all went off together in 1988 to start a pro
audio company. So, I have a certain place in my heart for the SRS
instruments of that era.
I will report back (sporadically?) as I work my way through this.
--Andrew
Andrew E. Kalman, Ph.D.
On Fri, Nov 4, 2022 at 3:22 PM Magnus Danielson via time-nuts <
time-nuts@lists.febo.com> wrote:
Hi Andrew,
On 11/4/22 19:44, Andrew Kalman via time-nuts wrote:
Hi All.
I have an SR620 that is misbehaving. I am working my way through the
manual
and schematics, but one consistent behavior has arisen: At a period of
around 4-6s when measuring frequency, I get a "wrong number" that is
consistent in scale across 10MHz and 1kHz (the REF frequency).
I.e. when I measure a 10MHz signal, occasionally the display shows
9946187.4... Hz and when measuring REF (1kHz) the display occasionally
shows 994.618757 Hz. This is independent of termination, and the A & B
front-ends have passed their tests.
Has anyone encountered the same issue, or have a suggestion as to where
to
look for the root cause?
[image: 20221104_SR620_SN356_Frequency_error-1-small.jpg]
[image: 20221104_SR620_SN356_Frequency_error-2-small.jpg]
[image: 20221104_SR620_SN356_Frequency_error-3-small.jpg]
Since f = cycles / time, these two latest is actually about the same time.
time = cycles / f and cycles is nominal frequency times nominal
time-base, so
t1 = 1E5 / 9946187.4882 = .0100541036 s
t2 = 1E1 / 994.618757 = .01005410357 s
So, both are really taking about the same time, just above the 10 ms the
time-base is set for, 54.1 us or so.
I would guess that there is an issue with the stop-counter side, so it
somehow keeps running. The difference you see is probably in the
different start-time-stamps.
The time as used in above formula is really stop-time minus start-time
to get the elapsed time, and the time-base triggers how far after the
start-time that an attempt at a stop-time is taken. So, it's not
unreasonable that we are just above 10 ms, it's the intended behavior,
but the actual values is consistent with some issue to it.
I hope this gives you a good clue of what part of the good old SR620
troubling you.
Sometimes I find that tweaking the trigger-points avoid these issues.
That is part of the working that turns into part of a skill. However,
when your trigger is in auto for normal good signals, you should not
experience this. If messing with your trigger-point help you, it could
be that you need to work on the trigger side of things to get stable
operation as default properly. The 1 kHz output on loopback should
always operate safely in auto-mode, yet you illustrate it's failure. I
seem to recall calibration details about it.
Cheers,
Magnus