volt-nuts@lists.febo.com

Discussion of precise voltage measurement

View all threads

Just got an Agilent 3458A

MY
Miguel Yepes
Mon, Jul 29, 2019 7:09 PM

The meter was more than 2 years in storage, what should I do with it? I have not even autocal I just turned on and passed the selftests. I was told the first thing is let it on for 2 months then send to calibration. I will use it sporadically to do verifications only. My serial number is: MY45044988.
Any other thing I should do to it?

The meter was more than 2 years in storage, what should I do with it? I have not even autocal I just turned on and passed the selftests. I was told the first thing is let it on for 2 months then send to calibration. I will use it sporadically to do verifications only. My serial number is: MY45044988. Any other thing I should do to it?
IT
Illya Tsemenko
Mon, Jul 29, 2019 7:13 PM

Sell? :) I'd suggest running SN18 service note test procedure for few
weeks, to figure out if ADC still okay, or you need $1.4K replacement
board...

On 7/30/2019 3:09 AM, Miguel Yepes wrote:

The meter was more than 2 years in storage, what should I do with it? I have not even autocal I just turned on and passed the selftests. I was told the first thing is let it on for 2 months then send to calibration. I will use it sporadically to do verifications only. My serial number is: MY45044988.
Any other thing I should do to it?


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

Sell? :) I'd suggest running SN18 service note test procedure for few weeks, to figure out if ADC still okay, or you need $1.4K replacement board... On 7/30/2019 3:09 AM, Miguel Yepes wrote: > The meter was more than 2 years in storage, what should I do with it? I have not even autocal I just turned on and passed the selftests. I was told the first thing is let it on for 2 months then send to calibration. I will use it sporadically to do verifications only. My serial number is: MY45044988. > Any other thing I should do to it? > _______________________________________________ > volt-nuts mailing list -- volt-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com > and follow the instructions there.
PK
Poul-Henning Kamp
Mon, Jul 29, 2019 8:33 PM

In message <CY4PR10MB1846411363993295BDCF2496C3DD0@CY4PR10MB1846.namprd10.prod.
outlook.com>, Miguel Yepes writes:

Any other thing I should do to it?

You should make a backup of the CALRAM.

I have some python code which does this:

https://github.com/bsdphk/pylt/blob/master/hp3458a.py

And somebody else has written a windows program that does it,
google or eevblog will find it for you.

Once you have made the backup, look at the date-code of the NVRAM
chip (the small one) and decide if you want to replace it before
or after the internal battery dies.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

-------- In message <CY4PR10MB1846411363993295BDCF2496C3DD0@CY4PR10MB1846.namprd10.prod. outlook.com>, Miguel Yepes writes: >Any other thing I should do to it? You should make a backup of the CALRAM. I have some python code which does this: https://github.com/bsdphk/pylt/blob/master/hp3458a.py And somebody else has written a windows program that does it, google or eevblog will find it for you. Once you have made the backup, look at the date-code of the NVRAM chip (the small one) and decide if you want to replace it before or after the internal battery dies. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
MY
Miguel Yepes
Mon, Jul 29, 2019 9:04 PM

But that doc it said from within these serial numbers:
Serial Numbers: US28031400 / US28032927
Mine is out of those.


From: volt-nuts volt-nuts-bounces@lists.febo.com on behalf of Illya Tsemenko illya@xdevs.com
Sent: Monday, July 29, 2019 2:13 PM
To: volt-nuts@lists.febo.com volt-nuts@lists.febo.com
Subject: Re: [volt-nuts] Just got an Agilent 3458A

Sell? :) I'd suggest running SN18 service note test procedure for few
weeks, to figure out if ADC still okay, or you need $1.4K replacement
board...

On 7/30/2019 3:09 AM, Miguel Yepes wrote:

The meter was more than 2 years in storage, what should I do with it? I have not even autocal I just turned on and passed the selftests. I was told the first thing is let it on for 2 months then send to calibration. I will use it sporadically to do verifications only. My serial number is: MY45044988.
Any other thing I should do to it?


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

But that doc it said from within these serial numbers: Serial Numbers: US28031400 / US28032927 Mine is out of those. ________________________________ From: volt-nuts <volt-nuts-bounces@lists.febo.com> on behalf of Illya Tsemenko <illya@xdevs.com> Sent: Monday, July 29, 2019 2:13 PM To: volt-nuts@lists.febo.com <volt-nuts@lists.febo.com> Subject: Re: [volt-nuts] Just got an Agilent 3458A Sell? :) I'd suggest running SN18 service note test procedure for few weeks, to figure out if ADC still okay, or you need $1.4K replacement board... On 7/30/2019 3:09 AM, Miguel Yepes wrote: > The meter was more than 2 years in storage, what should I do with it? I have not even autocal I just turned on and passed the selftests. I was told the first thing is let it on for 2 months then send to calibration. I will use it sporadically to do verifications only. My serial number is: MY45044988. > Any other thing I should do to it? > _______________________________________________ > volt-nuts mailing list -- volt-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com > and follow the instructions there. _______________________________________________ volt-nuts mailing list -- volt-nuts@lists.febo.com To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com and follow the instructions there.
TM
Todd Micallef
Mon, Jul 29, 2019 9:09 PM

Assume all used 3458A have been repaired or raided including ref boards.

One thing to check inside the meter are date codes on all the boards.

BR,
Todd

On Monday, July 29, 2019, Miguel Yepes theamberco@hotmail.com wrote:

But that doc it said from within these serial numbers:
Serial Numbers: US28031400 / US28032927
Mine is out of those.


From: volt-nuts volt-nuts-bounces@lists.febo.com on behalf of Illya
Tsemenko illya@xdevs.com
Sent: Monday, July 29, 2019 2:13 PM
To: volt-nuts@lists.febo.com volt-nuts@lists.febo.com
Subject: Re: [volt-nuts] Just got an Agilent 3458A

Sell? :) I'd suggest running SN18 service note test procedure for few
weeks, to figure out if ADC still okay, or you need $1.4K replacement
board...

On 7/30/2019 3:09 AM, Miguel Yepes wrote:

The meter was more than 2 years in storage, what should I do with it? I

have not even autocal I just turned on and passed the selftests. I was told
the first thing is let it on for 2 months then send to calibration. I will
use it sporadically to do verifications only. My serial number is:
MY45044988.

Any other thing I should do to it?


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/

listinfo/volt-nuts_lists.febo.com

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/
listinfo/volt-nuts_lists.febo.com
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/
listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

Assume all used 3458A have been repaired or raided including ref boards. One thing to check inside the meter are date codes on all the boards. BR, Todd On Monday, July 29, 2019, Miguel Yepes <theamberco@hotmail.com> wrote: > > But that doc it said from within these serial numbers: > Serial Numbers: US28031400 / US28032927 > Mine is out of those. > ________________________________ > From: volt-nuts <volt-nuts-bounces@lists.febo.com> on behalf of Illya > Tsemenko <illya@xdevs.com> > Sent: Monday, July 29, 2019 2:13 PM > To: volt-nuts@lists.febo.com <volt-nuts@lists.febo.com> > Subject: Re: [volt-nuts] Just got an Agilent 3458A > > Sell? :) I'd suggest running SN18 service note test procedure for few > weeks, to figure out if ADC still okay, or you need $1.4K replacement > board... > > On 7/30/2019 3:09 AM, Miguel Yepes wrote: > > The meter was more than 2 years in storage, what should I do with it? I > have not even autocal I just turned on and passed the selftests. I was told > the first thing is let it on for 2 months then send to calibration. I will > use it sporadically to do verifications only. My serial number is: > MY45044988. > > Any other thing I should do to it? > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@lists.febo.com > > To unsubscribe, go to http://lists.febo.com/mailman/ > listinfo/volt-nuts_lists.febo.com > > and follow the instructions there. > > > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/ > listinfo/volt-nuts_lists.febo.com > and follow the instructions there. > _______________________________________________ > volt-nuts mailing list -- volt-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/ > listinfo/volt-nuts_lists.febo.com > and follow the instructions there. >
MY
Miguel Yepes
Mon, Jul 29, 2019 9:13 PM

I was told this meter has not been repaired. It is in like new condition. It was bought for one job a few years back and them stored for over 2 years.


From: volt-nuts volt-nuts-bounces@lists.febo.com on behalf of Todd Micallef tmicallef@gmail.com
Sent: Monday, July 29, 2019 4:09 PM
To: Discussion of precise voltage measurement volt-nuts@lists.febo.com
Subject: Re: [volt-nuts] Just got an Agilent 3458A

Assume all used 3458A have been repaired or raided including ref boards.

One thing to check inside the meter are date codes on all the boards.

BR,
Todd

On Monday, July 29, 2019, Miguel Yepes theamberco@hotmail.com wrote:

But that doc it said from within these serial numbers:
Serial Numbers: US28031400 / US28032927
Mine is out of those.


From: volt-nuts volt-nuts-bounces@lists.febo.com on behalf of Illya
Tsemenko illya@xdevs.com
Sent: Monday, July 29, 2019 2:13 PM
To: volt-nuts@lists.febo.com volt-nuts@lists.febo.com
Subject: Re: [volt-nuts] Just got an Agilent 3458A

Sell? :) I'd suggest running SN18 service note test procedure for few
weeks, to figure out if ADC still okay, or you need $1.4K replacement
board...

On 7/30/2019 3:09 AM, Miguel Yepes wrote:

The meter was more than 2 years in storage, what should I do with it? I

have not even autocal I just turned on and passed the selftests. I was told
the first thing is let it on for 2 months then send to calibration. I will
use it sporadically to do verifications only. My serial number is:
MY45044988.

Any other thing I should do to it?


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/

listinfo/volt-nuts_lists.febo.com

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/
listinfo/volt-nuts_lists.febo.com
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/
listinfo/volt-nuts_lists.febo.com
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

I was told this meter has not been repaired. It is in like new condition. It was bought for one job a few years back and them stored for over 2 years. ________________________________ From: volt-nuts <volt-nuts-bounces@lists.febo.com> on behalf of Todd Micallef <tmicallef@gmail.com> Sent: Monday, July 29, 2019 4:09 PM To: Discussion of precise voltage measurement <volt-nuts@lists.febo.com> Subject: Re: [volt-nuts] Just got an Agilent 3458A Assume all used 3458A have been repaired or raided including ref boards. One thing to check inside the meter are date codes on all the boards. BR, Todd On Monday, July 29, 2019, Miguel Yepes <theamberco@hotmail.com> wrote: > > But that doc it said from within these serial numbers: > Serial Numbers: US28031400 / US28032927 > Mine is out of those. > ________________________________ > From: volt-nuts <volt-nuts-bounces@lists.febo.com> on behalf of Illya > Tsemenko <illya@xdevs.com> > Sent: Monday, July 29, 2019 2:13 PM > To: volt-nuts@lists.febo.com <volt-nuts@lists.febo.com> > Subject: Re: [volt-nuts] Just got an Agilent 3458A > > Sell? :) I'd suggest running SN18 service note test procedure for few > weeks, to figure out if ADC still okay, or you need $1.4K replacement > board... > > On 7/30/2019 3:09 AM, Miguel Yepes wrote: > > The meter was more than 2 years in storage, what should I do with it? I > have not even autocal I just turned on and passed the selftests. I was told > the first thing is let it on for 2 months then send to calibration. I will > use it sporadically to do verifications only. My serial number is: > MY45044988. > > Any other thing I should do to it? > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@lists.febo.com > > To unsubscribe, go to http://lists.febo.com/mailman/ > listinfo/volt-nuts_lists.febo.com > > and follow the instructions there. > > > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/ > listinfo/volt-nuts_lists.febo.com > and follow the instructions there. > _______________________________________________ > volt-nuts mailing list -- volt-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/ > listinfo/volt-nuts_lists.febo.com > and follow the instructions there. > _______________________________________________ volt-nuts mailing list -- volt-nuts@lists.febo.com To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com and follow the instructions there.
DD
Dr. David Kirkby
Mon, Jul 29, 2019 9:37 PM

On Mon, 29 Jul 2019 at 22:13, Miguel Yepes theamberco@hotmail.com wrote:

I was told this meter has not been repaired. It is in like new condition.
It was bought for one job a few years back and them stored for over 2 years.

It must have been bought second-hand if bought a few years back. Agilent
was split off from HP in 1999. So whilst 1999 is not the latest year a
meter could be badged HP, and still sold by Agilent, it would probably be
2001 - not all labels were changed overnight. I believe there was a period
of a couple of years, where something could be badged HP, but still sold by
Agilent.  So that puts it at least 18 years old.

There is a rule,

https://www.eevblog.com/forum/testgear/agilent-hp-serial-number-confusion/

which I don't know if it's 100% reliable, that the year of the last design
change is 1960 + the first two digits of the serial number (MY45044988). If
that was so, the last design change was 1960 + 45 = 2005. But if the design
was changed in 2005, that would have been 6 years after HP was split off
into Agilent, so it would be badged Agilent, not HP.

I suspect that the meter is made from at least two meters, which seems to
be quite common on 3458As.

Dr David Kirkby Ph.D C.Eng MIET
Kirkby Microwave Ltd
Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, CHELMSFORD,
Essex, CM3 6DT, United Kingdom.
Registered in England and Wales as company number 08914892
https://www.kirkbymicrowave.co.uk/
Tel 01621-680100 / +44 1621-680100

On Mon, 29 Jul 2019 at 22:13, Miguel Yepes <theamberco@hotmail.com> wrote: > I was told this meter has not been repaired. It is in like new condition. > It was bought for one job a few years back and them stored for over 2 years. > It must have been bought second-hand if bought a few years back. Agilent was split off from HP in 1999. So whilst 1999 is not the latest year a meter could be badged HP, and still sold by Agilent, it would probably be 2001 - not all labels were changed overnight. I believe there was a period of a couple of years, where something could be badged HP, but still sold by Agilent. So that puts it at least 18 years old. There is a rule, https://www.eevblog.com/forum/testgear/agilent-hp-serial-number-confusion/ which I don't know if it's 100% reliable, that the year of the last design change is 1960 + the first two digits of the serial number (MY45044988). If that was so, the last design change was 1960 + 45 = 2005. But if the design was changed in 2005, that would have been 6 years after HP was split off into Agilent, so it would be badged Agilent, not HP. I suspect that the meter is made from at least two meters, which seems to be quite common on 3458As. Dr David Kirkby Ph.D C.Eng MIET Kirkby Microwave Ltd Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, CHELMSFORD, Essex, CM3 6DT, United Kingdom. Registered in England and Wales as company number 08914892 https://www.kirkbymicrowave.co.uk/ Tel 01621-680100 / +44 1621-680100
MY
Miguel Yepes
Mon, Jul 29, 2019 10:59 PM

No idea what are you talking about my meter is branded Agilent and by the looks of it is one of the lastest ones, I need to check the serial to know for sure.


From: volt-nuts volt-nuts-bounces@lists.febo.com on behalf of Dr. David Kirkby drkirkby@kirkbymicrowave.co.uk
Sent: Monday, July 29, 2019 4:37 PM
To: Discussion of precise voltage measurement volt-nuts@lists.febo.com
Subject: Re: [volt-nuts] Just got an Agilent 3458A

On Mon, 29 Jul 2019 at 22:13, Miguel Yepes theamberco@hotmail.com wrote:

I was told this meter has not been repaired. It is in like new condition.
It was bought for one job a few years back and them stored for over 2 years.

It must have been bought second-hand if bought a few years back. Agilent
was split off from HP in 1999. So whilst 1999 is not the latest year a
meter could be badged HP, and still sold by Agilent, it would probably be
2001 - not all labels were changed overnight. I believe there was a period
of a couple of years, where something could be badged HP, but still sold by
Agilent.  So that puts it at least 18 years old.

There is a rule,

https://www.eevblog.com/forum/testgear/agilent-hp-serial-number-confusion/

which I don't know if it's 100% reliable, that the year of the last design
change is 1960 + the first two digits of the serial number (MY45044988). If
that was so, the last design change was 1960 + 45 = 2005. But if the design
was changed in 2005, that would have been 6 years after HP was split off
into Agilent, so it would be badged Agilent, not HP.

I suspect that the meter is made from at least two meters, which seems to
be quite common on 3458As.

Dr David Kirkby Ph.D C.Eng MIET
Kirkby Microwave Ltd
Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, CHELMSFORD,
Essex, CM3 6DT, United Kingdom.
Registered in England and Wales as company number 08914892
https://www.kirkbymicrowave.co.uk/
Tel 01621-680100 / +44 1621-680100


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

No idea what are you talking about my meter is branded Agilent and by the looks of it is one of the lastest ones, I need to check the serial to know for sure. ________________________________ From: volt-nuts <volt-nuts-bounces@lists.febo.com> on behalf of Dr. David Kirkby <drkirkby@kirkbymicrowave.co.uk> Sent: Monday, July 29, 2019 4:37 PM To: Discussion of precise voltage measurement <volt-nuts@lists.febo.com> Subject: Re: [volt-nuts] Just got an Agilent 3458A On Mon, 29 Jul 2019 at 22:13, Miguel Yepes <theamberco@hotmail.com> wrote: > I was told this meter has not been repaired. It is in like new condition. > It was bought for one job a few years back and them stored for over 2 years. > It must have been bought second-hand if bought a few years back. Agilent was split off from HP in 1999. So whilst 1999 is not the latest year a meter could be badged HP, and still sold by Agilent, it would probably be 2001 - not all labels were changed overnight. I believe there was a period of a couple of years, where something could be badged HP, but still sold by Agilent. So that puts it at least 18 years old. There is a rule, https://www.eevblog.com/forum/testgear/agilent-hp-serial-number-confusion/ which I don't know if it's 100% reliable, that the year of the last design change is 1960 + the first two digits of the serial number (MY45044988). If that was so, the last design change was 1960 + 45 = 2005. But if the design was changed in 2005, that would have been 6 years after HP was split off into Agilent, so it would be badged Agilent, not HP. I suspect that the meter is made from at least two meters, which seems to be quite common on 3458As. Dr David Kirkby Ph.D C.Eng MIET Kirkby Microwave Ltd Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, CHELMSFORD, Essex, CM3 6DT, United Kingdom. Registered in England and Wales as company number 08914892 https://www.kirkbymicrowave.co.uk/ Tel 01621-680100 / +44 1621-680100 _______________________________________________ volt-nuts mailing list -- volt-nuts@lists.febo.com To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com and follow the instructions there.
JL
J. L. Trantham
Mon, Jul 29, 2019 11:08 PM

The 3458A is still supported by Keysight.  You can get a Keysight
Calibration for $777.

If it passes Calibration, then you are eligible to purchase a Keysight
Repair Agreement for $216 per year.  You can (or at least you could) buy up
to a 5 year agreement with a 5% (IIRC) discount.  As opposed to a 'per
incident' repair of $3099.  Or at least as listed at Keysight US Website.

I've done this with my 3458A's and will be investigating them extensively if
I ever retire.

Good luck.

Joe

-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@lists.febo.com] On Behalf Of Dr.
David Kirkby
Sent: Monday, July 29, 2019 4:37 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Just got an Agilent 3458A

On Mon, 29 Jul 2019 at 22:13, Miguel Yepes theamberco@hotmail.com wrote:

I was told this meter has not been repaired. It is in like new condition.
It was bought for one job a few years back and them stored for over 2

years.

It must have been bought second-hand if bought a few years back. Agilent
was split off from HP in 1999. So whilst 1999 is not the latest year a
meter could be badged HP, and still sold by Agilent, it would probably be
2001 - not all labels were changed overnight. I believe there was a period
of a couple of years, where something could be badged HP, but still sold by
Agilent.  So that puts it at least 18 years old.

There is a rule,

https://www.eevblog.com/forum/testgear/agilent-hp-serial-number-confusion/

which I don't know if it's 100% reliable, that the year of the last design
change is 1960 + the first two digits of the serial number (MY45044988). If
that was so, the last design change was 1960 + 45 = 2005. But if the design
was changed in 2005, that would have been 6 years after HP was split off
into Agilent, so it would be badged Agilent, not HP.

I suspect that the meter is made from at least two meters, which seems to
be quite common on 3458As.

Dr David Kirkby Ph.D C.Eng MIET
Kirkby Microwave Ltd
Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, CHELMSFORD,
Essex, CM3 6DT, United Kingdom.
Registered in England and Wales as company number 08914892
https://www.kirkbymicrowave.co.uk/
Tel 01621-680100 / +44 1621-680100


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

The 3458A is still supported by Keysight. You can get a Keysight Calibration for $777. If it passes Calibration, then you are eligible to purchase a Keysight Repair Agreement for $216 per year. You can (or at least you could) buy up to a 5 year agreement with a 5% (IIRC) discount. As opposed to a 'per incident' repair of $3099. Or at least as listed at Keysight US Website. I've done this with my 3458A's and will be investigating them extensively if I ever retire. Good luck. Joe -----Original Message----- From: volt-nuts [mailto:volt-nuts-bounces@lists.febo.com] On Behalf Of Dr. David Kirkby Sent: Monday, July 29, 2019 4:37 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Just got an Agilent 3458A On Mon, 29 Jul 2019 at 22:13, Miguel Yepes <theamberco@hotmail.com> wrote: > I was told this meter has not been repaired. It is in like new condition. > It was bought for one job a few years back and them stored for over 2 years. > It must have been bought second-hand if bought a few years back. Agilent was split off from HP in 1999. So whilst 1999 is not the latest year a meter could be badged HP, and still sold by Agilent, it would probably be 2001 - not all labels were changed overnight. I believe there was a period of a couple of years, where something could be badged HP, but still sold by Agilent. So that puts it at least 18 years old. There is a rule, https://www.eevblog.com/forum/testgear/agilent-hp-serial-number-confusion/ which I don't know if it's 100% reliable, that the year of the last design change is 1960 + the first two digits of the serial number (MY45044988). If that was so, the last design change was 1960 + 45 = 2005. But if the design was changed in 2005, that would have been 6 years after HP was split off into Agilent, so it would be badged Agilent, not HP. I suspect that the meter is made from at least two meters, which seems to be quite common on 3458As. Dr David Kirkby Ph.D C.Eng MIET Kirkby Microwave Ltd Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, CHELMSFORD, Essex, CM3 6DT, United Kingdom. Registered in England and Wales as company number 08914892 https://www.kirkbymicrowave.co.uk/ Tel 01621-680100 / +44 1621-680100 _______________________________________________ volt-nuts mailing list -- volt-nuts@lists.febo.com To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com and follow the instructions there.
DD
Dr. David Kirkby
Tue, Jul 30, 2019 4:11 AM

On Tue, 30 Jul 2019 at 00:00, Miguel Yepes theamberco@hotmail.com wrote:

No idea what are you talking about my meter is branded Agilent and by the
looks of it is one of the lastest ones, I need to check the serial to know
for sure.

Oh, my error. I thought the photograph you emailed me said HP. A bad memory
on my part.

it would appear that the latest engineering change to yours was 2005, so 14
years old. I have no idea if there are engineering changes since then.

Dave.


From: volt-nuts volt-nuts-bounces@lists.febo.com on behalf of Dr. David
Kirkby drkirkby@kirkbymicrowave.co.uk
Sent: Monday, July 29, 2019 4:37 PM
To: Discussion of precise voltage measurement volt-nuts@lists.febo.com
Subject: Re: [volt-nuts] Just got an Agilent 3458A

On Mon, 29 Jul 2019 at 22:13, Miguel Yepes theamberco@hotmail.com wrote:

I was told this meter has not been repaired. It is in like new condition.
It was bought for one job a few years back and them stored for over 2

years.

It must have been bought second-hand if bought a few years back. Agilent
was split off from HP in 1999. So whilst 1999 is not the latest year a
meter could be badged HP, and still sold by Agilent, it would probably be
2001 - not all labels were changed overnight. I believe there was a period
of a couple of years, where something could be badged HP, but still sold by
Agilent.  So that puts it at least 18 years old.

There is a rule,

https://www.eevblog.com/forum/testgear/agilent-hp-serial-number-confusion/

which I don't know if it's 100% reliable, that the year of the last design
change is 1960 + the first two digits of the serial number (MY45044988). If
that was so, the last design change was 1960 + 45 = 2005. But if the design
was changed in 2005, that would have been 6 years after HP was split off
into Agilent, so it would be badged Agilent, not HP.

I suspect that the meter is made from at least two meters, which seems to
be quite common on 3458As.

Dr David Kirkby Ph.D C.Eng MIET
Kirkby Microwave Ltd
Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, CHELMSFORD,
Essex, CM3 6DT, United Kingdom.
Registered in England and Wales as company number 08914892
https://www.kirkbymicrowave.co.uk/
Tel 01621-680100 / +44 1621-680100


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

--
Dr. David Kirkby,

On Tue, 30 Jul 2019 at 00:00, Miguel Yepes <theamberco@hotmail.com> wrote: > No idea what are you talking about my meter is branded Agilent and by the > looks of it is one of the lastest ones, I need to check the serial to know > for sure. Oh, my error. I thought the photograph you emailed me said HP. A bad memory on my part. it would appear that the latest engineering change to yours was 2005, so 14 years old. I have no idea if there are engineering changes since then. Dave. > > > ________________________________ > From: volt-nuts <volt-nuts-bounces@lists.febo.com> on behalf of Dr. David > Kirkby <drkirkby@kirkbymicrowave.co.uk> > Sent: Monday, July 29, 2019 4:37 PM > To: Discussion of precise voltage measurement <volt-nuts@lists.febo.com> > Subject: Re: [volt-nuts] Just got an Agilent 3458A > > On Mon, 29 Jul 2019 at 22:13, Miguel Yepes <theamberco@hotmail.com> wrote: > > > I was told this meter has not been repaired. It is in like new condition. > > It was bought for one job a few years back and them stored for over 2 > years. > > > > It must have been bought second-hand if bought a few years back. Agilent > was split off from HP in 1999. So whilst 1999 is not the latest year a > meter could be badged HP, and still sold by Agilent, it would probably be > 2001 - not all labels were changed overnight. I believe there was a period > of a couple of years, where something could be badged HP, but still sold by > Agilent. So that puts it at least 18 years old. > > There is a rule, > > https://www.eevblog.com/forum/testgear/agilent-hp-serial-number-confusion/ > > which I don't know if it's 100% reliable, that the year of the last design > change is 1960 + the first two digits of the serial number (MY45044988). If > that was so, the last design change was 1960 + 45 = 2005. But if the design > was changed in 2005, that would have been 6 years after HP was split off > into Agilent, so it would be badged Agilent, not HP. > > I suspect that the meter is made from at least two meters, which seems to > be quite common on 3458As. > > Dr David Kirkby Ph.D C.Eng MIET > Kirkby Microwave Ltd > Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, CHELMSFORD, > Essex, CM3 6DT, United Kingdom. > Registered in England and Wales as company number 08914892 > https://www.kirkbymicrowave.co.uk/ > Tel 01621-680100 / +44 1621-680100 > _______________________________________________ > volt-nuts mailing list -- volt-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com > and follow the instructions there. > _______________________________________________ > volt-nuts mailing list -- volt-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com > and follow the instructions there. > -- Dr. David Kirkby,
DD
Dr. David Kirkby
Tue, Jul 30, 2019 4:38 AM

On Tue, 30 Jul 2019 at 00:00, Miguel Yepes theamberco@hotmail.com wrote:

No idea what are you talking about my meter is branded Agilent and by the
looks of it is one of the lastest ones, I need to check the serial to know
for sure.

Correction.

The S/N indicates that the last engineering change was in 2005, which is 14
years ago. I have no idea if the design has changed since then, but others
may know.

Keysight was formed in 2014, which is 5 years ago. I would guess that as
late as 2016, things could still be badged Agilent. So it could be as as
little as 3 years old.

That puts it somewhere between 3 and 14 tears old.

If you put high definition photos of each board on a website, people on
here might be able to date it better.

Dave

Dr. David Kirkby,

On Tue, 30 Jul 2019 at 00:00, Miguel Yepes <theamberco@hotmail.com> wrote: > No idea what are you talking about my meter is branded Agilent and by the > looks of it is one of the lastest ones, I need to check the serial to know > for sure. Correction. The S/N indicates that the last engineering change was in 2005, which is 14 years ago. I have no idea if the design has changed since then, but others may know. Keysight was formed in 2014, which is 5 years ago. I would guess that as late as 2016, things could still be badged Agilent. So it could be as as little as 3 years old. That puts it somewhere between 3 and 14 tears old. If you put high definition photos of each board on a website, people on here might be able to date it better. Dave -- Dr. David Kirkby,
DD
Dr. David Kirkby
Tue, Jul 30, 2019 4:45 AM

On Tue, 30 Jul 2019 at 00:09, J. L. Trantham jltran@att.net wrote:

The 3458A is still supported by Keysight.  You can get a Keysight
Calibration for $777.

If it passes Calibration, then you are eligible to purchase a Keysight
Repair Agreement for $216 per year.  You can (or at least you could) buy up
to a 5 year agreement with a 5% (IIRC) discount.  As opposed to a 'per
incident' repair of $3099.  Or at least as listed at Keysight US Website.

That seems good value, given repairs can be expensive.

Is there any requirement for it to be sent for calibration each year during
that period?

If it has been in storage a couple of years I guess one should run it
continuously a while before sending for calibration. But how long?

Dave

Dr. David Kirkby,

On Tue, 30 Jul 2019 at 00:09, J. L. Trantham <jltran@att.net> wrote: > The 3458A is still supported by Keysight. You can get a Keysight > Calibration for $777. > > If it passes Calibration, then you are eligible to purchase a Keysight > Repair Agreement for $216 per year. You can (or at least you could) buy up > to a 5 year agreement with a 5% (IIRC) discount. As opposed to a 'per > incident' repair of $3099. Or at least as listed at Keysight US Website. That seems good value, given repairs can be expensive. Is there any requirement for it to be sent for calibration each year during that period? If it has been in storage a couple of years I guess one should run it continuously a while before sending for calibration. But how long? Dave -- Dr. David Kirkby,
JL
J. L. Trantham
Tue, Jul 30, 2019 12:28 PM

As long as it's under the Repair Agreement, you're covered.  In order to get
the Repair Agreement, it must be able to be calibrated.

I've purchased 5 year Agreements so I send it in for calibration every 3 to
5 years.

Joe

-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@lists.febo.com] On Behalf Of Dr.
David Kirkby
Sent: Monday, July 29, 2019 11:45 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Just got an Agilent 3458A

On Tue, 30 Jul 2019 at 00:09, J. L. Trantham jltran@att.net wrote:

The 3458A is still supported by Keysight.  You can get a Keysight
Calibration for $777.

If it passes Calibration, then you are eligible to purchase a Keysight
Repair Agreement for $216 per year.  You can (or at least you could) buy

up

to a 5 year agreement with a 5% (IIRC) discount.  As opposed to a 'per
incident' repair of $3099.  Or at least as listed at Keysight US Website.

That seems good value, given repairs can be expensive.

Is there any requirement for it to be sent for calibration each year during
that period?

If it has been in storage a couple of years I guess one should run it
continuously a while before sending for calibration. But how long?

Dave

Dr. David Kirkby,


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

As long as it's under the Repair Agreement, you're covered. In order to get the Repair Agreement, it must be able to be calibrated. I've purchased 5 year Agreements so I send it in for calibration every 3 to 5 years. Joe -----Original Message----- From: volt-nuts [mailto:volt-nuts-bounces@lists.febo.com] On Behalf Of Dr. David Kirkby Sent: Monday, July 29, 2019 11:45 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Just got an Agilent 3458A On Tue, 30 Jul 2019 at 00:09, J. L. Trantham <jltran@att.net> wrote: > The 3458A is still supported by Keysight. You can get a Keysight > Calibration for $777. > > If it passes Calibration, then you are eligible to purchase a Keysight > Repair Agreement for $216 per year. You can (or at least you could) buy up > to a 5 year agreement with a 5% (IIRC) discount. As opposed to a 'per > incident' repair of $3099. Or at least as listed at Keysight US Website. That seems good value, given repairs can be expensive. Is there any requirement for it to be sent for calibration each year during that period? If it has been in storage a couple of years I guess one should run it continuously a while before sending for calibration. But how long? Dave -- Dr. David Kirkby, _______________________________________________ volt-nuts mailing list -- volt-nuts@lists.febo.com To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com and follow the instructions there.