Validating GPSDO control loop with simulation and measurement. Is this amount of difference between measurement and simulation to be expected?

EK
Erik Kaashoek
Fri, Feb 25, 2022 3:18 PM

Inspired by the gpsdo simulator written by Tom Van Baak  I am trying to
use simulation to validate the PID loop parameters for a cheap and
simple GPS referenced timer/counter I'm building.
The following one hour measurements where done using a Rbd reference as
input to the timer counter(see Timelab.gif plot attached)
1: Open loop (P=0, I=0) VC-TCXO versus the Rbd reference (green trace)
2: PPS of the internal GPS versus the Rbd reference. (PPS trace)
3: Rbd reference versus the closed loop (P=0.02, I=0) VC-TCXO (dark blue
trace)
4: Rbd reference versus the closed loop (P=0.01, I=0,00005) VC-TCXO (red
trace)
All .tim files also attached.
The performance of the TCXO was remarkably good for a sub 1$ device, 
great care went into  a stable, low noise, Vtune. The noisy supply did
not have a big influence
The open loop TCXO frequencies where divided by 1e7 (SW PICdiv)and
together with the raw PPS frequencies loaded into an Excel speadsheet
(see attached)
An identical PID controller  (P=0.01, I=0.0005) was implemented in Excel
that took as input the frequency difference between the PPS and the TCXO
and calculated a correction for the TCXO frequency. The corrected TCXO
frequencies was exported from excel (PID output) and imported in Timelab
(PID, light blue trace)
As can be seen in the Timelab plot the controller implemented in Excel
did a better job then the controller in the timer/counter but the two
actual closed loop measurements used difference GPS and TCXO data (I do
not own sufficient good frequency counters to measure all frequencies at
the same time).
The timer/counter was open on a table with minimum thermal isolation and
during the measurements some draft through door opening/closing could occur
My question: is the difference between the measured loop performance and
the simulated  loop performance to be expected or am I making a big
mistake somewhere?
Feel free to ask more info or more measurement data

Erik.

Inspired by the gpsdo simulator written by Tom Van Baak  I am trying to use simulation to validate the PID loop parameters for a cheap and simple GPS referenced timer/counter I'm building. The following one hour measurements where done using a Rbd reference as input to the timer counter(see Timelab.gif plot attached) 1: Open loop (P=0, I=0) VC-TCXO versus the Rbd reference (green trace) 2: PPS of the internal GPS versus the Rbd reference. (PPS trace) 3: Rbd reference versus the closed loop (P=0.02, I=0) VC-TCXO (dark blue trace) 4: Rbd reference versus the closed loop (P=0.01, I=0,00005) VC-TCXO (red trace) All .tim files also attached. The performance of the TCXO was remarkably good for a sub 1$ device,  great care went into  a stable, low noise, Vtune. The noisy supply did not have a big influence The open loop TCXO frequencies where divided by 1e7 (SW PICdiv)and together with the raw PPS frequencies loaded into an Excel speadsheet (see attached) An identical PID controller  (P=0.01, I=0.0005) was implemented in Excel that took as input the frequency difference between the PPS and the TCXO and calculated a correction for the TCXO frequency. The corrected TCXO frequencies was exported from excel (PID output) and imported in Timelab (PID, light blue trace) As can be seen in the Timelab plot the controller implemented in Excel did a better job then the controller in the timer/counter but the two actual closed loop measurements used difference GPS and TCXO data (I do not own sufficient good frequency counters to measure all frequencies at the same time). The timer/counter was open on a table with minimum thermal isolation and during the measurements some draft through door opening/closing could occur My question: is the difference between the measured loop performance and the simulated  loop performance to be expected or am I making a big mistake somewhere? Feel free to ask more info or more measurement data Erik.
BK
Bob kb8tq
Fri, Feb 25, 2022 5:22 PM

Hi

On Feb 25, 2022, at 10:18 AM, Erik Kaashoek erik@kaashoek.com wrote:

Inspired by the gpsdo simulator written by Tom Van Baak  I am trying to use simulation to validate the PID loop parameters for a cheap and simple GPS referenced timer/counter I'm building.
The following one hour measurements where done using a Rbd reference as input to the timer counter(see Timelab.gif plot attached)
1: Open loop (P=0, I=0) VC-TCXO versus the Rbd reference (green trace)
2: PPS of the internal GPS versus the Rbd reference. (PPS trace)
3: Rbd reference versus the closed loop (P=0.02, I=0) VC-TCXO (dark blue trace)
4: Rbd reference versus the closed loop (P=0.01, I=0,00005) VC-TCXO (red trace)
All .tim files also attached.
The performance of the TCXO was remarkably good for a sub 1$ device,  great care went into  a stable, low noise, Vtune. The noisy supply did not have a big influence
The open loop TCXO frequencies where divided by 1e7 (SW PICdiv)and together with the raw PPS frequencies loaded into an Excel speadsheet (see attached)
An identical PID controller  (P=0.01, I=0.0005) was implemented in Excel that took as input the frequency difference between the PPS and the TCXO and calculated a correction for the TCXO frequency. The corrected TCXO frequencies was exported from excel (PID output) and imported in Timelab (PID, light blue trace)
As can be seen in the Timelab plot the controller implemented in Excel did a better job then the controller in the timer/counter but the two actual closed loop measurements used difference GPS and TCXO data (I do not own sufficient good frequency counters to measure all frequencies at the same time).
The timer/counter was open on a table with minimum thermal isolation and during the measurements some draft through door opening/closing could occur
My question: is the difference between the measured loop performance and the simulated  loop performance to be expected or am I making a big mistake somewhere?

The TCXO should converge to match the PPS. It should do that at some point past
the few hundred seconds range on your loop. You would need a longer data run to
really know if this is happening or not. Something > 10 hours ( so overnight works well …)
would be better.

Another thing to look into: You can likely improve the performance of the PPS by 4X
or more by using the sawtooth correction process. That would shift the whole graph
to the left and turn a couple hundred seconds into sorta kinda a hundred seconds.
Overnight runs would then become closer to a couple hours.

Another check on things: The PPS should roughly drop off as 1/tau. It will depart from
this once you get out a ways. Over the range you are looking at, that’s a pretty good
approximation. Wobbles in the line might indicate some sort of gotcha in the data collection
or (maybe) issues somewhere else.

Bob

Feel free to ask more info or more measurement data

Erik.

<Timelab.gif><PID.xlsx><TCXO.tim><PPS.tim><TCXO GPS locked P=0.02.tim><TCXO GPS locked P=0.01, I=0.00005.tim><PID output.tim>_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

Hi > On Feb 25, 2022, at 10:18 AM, Erik Kaashoek <erik@kaashoek.com> wrote: > > Inspired by the gpsdo simulator written by Tom Van Baak I am trying to use simulation to validate the PID loop parameters for a cheap and simple GPS referenced timer/counter I'm building. > The following one hour measurements where done using a Rbd reference as input to the timer counter(see Timelab.gif plot attached) > 1: Open loop (P=0, I=0) VC-TCXO versus the Rbd reference (green trace) > 2: PPS of the internal GPS versus the Rbd reference. (PPS trace) > 3: Rbd reference versus the closed loop (P=0.02, I=0) VC-TCXO (dark blue trace) > 4: Rbd reference versus the closed loop (P=0.01, I=0,00005) VC-TCXO (red trace) > All .tim files also attached. > The performance of the TCXO was remarkably good for a sub 1$ device, great care went into a stable, low noise, Vtune. The noisy supply did not have a big influence > The open loop TCXO frequencies where divided by 1e7 (SW PICdiv)and together with the raw PPS frequencies loaded into an Excel speadsheet (see attached) > An identical PID controller (P=0.01, I=0.0005) was implemented in Excel that took as input the frequency difference between the PPS and the TCXO and calculated a correction for the TCXO frequency. The corrected TCXO frequencies was exported from excel (PID output) and imported in Timelab (PID, light blue trace) > As can be seen in the Timelab plot the controller implemented in Excel did a better job then the controller in the timer/counter but the two actual closed loop measurements used difference GPS and TCXO data (I do not own sufficient good frequency counters to measure all frequencies at the same time). > The timer/counter was open on a table with minimum thermal isolation and during the measurements some draft through door opening/closing could occur > My question: is the difference between the measured loop performance and the simulated loop performance to be expected or am I making a big mistake somewhere? The TCXO *should* converge to match the PPS. It should do that at some point past the few hundred seconds range on your loop. You would need a longer data run to really know if this is happening or not. Something > 10 hours ( so overnight works well …) would be better. Another thing to look into: You can likely improve the performance of the PPS by 4X or more by using the sawtooth correction process. That would shift the whole graph to the left and turn a couple hundred seconds into sorta kinda a hundred seconds. Overnight runs would then become closer to a couple hours. Another check on things: The PPS should roughly drop off as 1/tau. It will depart from this once you get out a ways. Over the range you are looking at, that’s a pretty good approximation. Wobbles in the line might indicate some sort of gotcha in the data collection or (maybe) issues somewhere else. Bob > Feel free to ask more info or more measurement data > > Erik. > > > > > > <Timelab.gif><PID.xlsx><TCXO.tim><PPS.tim><TCXO GPS locked P=0.02.tim><TCXO GPS locked P=0.01, I=0.00005.tim><PID output.tim>_______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there.
MD
Magnus Danielson
Fri, Feb 25, 2022 10:18 PM

Hi Erik,

A few comments and suggestions on the way.

Please note that for a PLL, a PID is equivalent to a PI regulator. Turns
out that the P and D factors of the PID sums to form the P of the PI
regulator, while the I transfer right over. I have seen no benefit in
real implementation of adding the D factor, it did not add any magic to
convergence.

Also note that P is proportional to f0*d and I is proportional to f0^2
where f0 is the PLL resonance-frequency/cut-over-frequency and d is the
damping factor. This is trival to derive and I usually do it on a white
board or a single sheet of  A4 paper. I strongly advice you to use f0
and d as your steering parameters rather than P and I, it will make the
work better. Also, I recommend to keep d at least to 3 to avoid jitter
peaking issues.

The actual formulas contain a little more details with a few constants
here and there, but those can be lumped into a single aggregate constant
for each of the P and I, and knowing the basic relationship allows for
quick test and tweak. I've taught this to several colleagues and they
ended up testing their way to production values without too much effort,
they simply tweak them until they are satisfied with all the tests readings.

The gpsdo simulator tool is very fun and educational to learn from. I've
written several similar simulation tools and they have proven very
useful. I recommend to do tests to validate that the model and the
actual implementation match up.

One technique to decide f0 (or it's reciprocal the timeconstant) is to
plot the phase-noise of the reference and the controlled oscillator and
simply choose f0 to be at the intercept point of the two graphs. A
similar approach have been used on ADEV plots, in which case it is
called the Allan intercept point. The Allan intercept point is maybe a
little less well founded than the phase-noise variants, but try both and
see how you end up, it will be a learning experience.

For a PI PLL, you can usually test overshot reaction to a phase or
frequency step. The amount of overshot is directly related to the actual
damping factor, so that is easy to validate. Further, forcing it into a
low damping factor, you get a high Q so if you stress the loop with a
phase or frequency step, the period of the resulting ringing will
disclose the actual f0 of the loop. Using these two measurements, you
can fairly trivially establish the actual constants of this equation system:

P = Kp * f0 * d
I = Ki * f0^2

Anyway, knowing the actual f0 and d for your implementation and being
able to control those through the above equations will be of immense
help as you then aim to optimize performance.

Notice that scaling factor includes EFC input sensitivity, so if you
have different oscillators, break out that as the Ko factor and set it
properly for each oscillator. It is actually fairly simple to measure
the actual sensitivity by intentionally steer the EFC and see how much
frequency change, and you will trivially calculate the steering factor
Ko. With a little bit of work, one can make it self-calibrate this for
the oscillator.

As for your question, for your simulation and your implementation to
match up, you need to validate that your model of your implementation
match up and have know scale factors. Spending time to make sure things
work as you expect then saves you tons of work and you can use this
knowledge to simulate and then verify in actual implementation (which
tends to be slow) only a few times.

Cheers,
Magnus

On 2022-02-25 16:18, Erik Kaashoek wrote:

Inspired by the gpsdo simulator written by Tom Van Baak  I am trying
to use simulation to validate the PID loop parameters for a cheap and
simple GPS referenced timer/counter I'm building.
The following one hour measurements where done using a Rbd reference
as input to the timer counter(see Timelab.gif plot attached)
1: Open loop (P=0, I=0) VC-TCXO versus the Rbd reference (green trace)
2: PPS of the internal GPS versus the Rbd reference. (PPS trace)
3: Rbd reference versus the closed loop (P=0.02, I=0) VC-TCXO (dark
blue trace)
4: Rbd reference versus the closed loop (P=0.01, I=0,00005) VC-TCXO
(red trace)
All .tim files also attached.
The performance of the TCXO was remarkably good for a sub 1$ device, 
great care went into  a stable, low noise, Vtune. The noisy supply did
not have a big influence
The open loop TCXO frequencies where divided by 1e7 (SW PICdiv)and
together with the raw PPS frequencies loaded into an Excel speadsheet
(see attached)
An identical PID controller  (P=0.01, I=0.0005) was implemented in
Excel that took as input the frequency difference between the PPS and
the TCXO and calculated a correction for the TCXO frequency. The
corrected TCXO frequencies was exported from excel (PID output) and
imported in Timelab (PID, light blue trace)
As can be seen in the Timelab plot the controller implemented in Excel
did a better job then the controller in the timer/counter but the two
actual closed loop measurements used difference GPS and TCXO data (I
do not own sufficient good frequency counters to measure all
frequencies at the same time).
The timer/counter was open on a table with minimum thermal isolation
and during the measurements some draft through door opening/closing
could occur
My question: is the difference between the measured loop performance
and the simulated  loop performance to be expected or am I making a
big mistake somewhere?
Feel free to ask more info or more measurement data

Erik.


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

Hi Erik, A few comments and suggestions on the way. Please note that for a PLL, a PID is equivalent to a PI regulator. Turns out that the P and D factors of the PID sums to form the P of the PI regulator, while the I transfer right over. I have seen no benefit in real implementation of adding the D factor, it did not add any magic to convergence. Also note that P is proportional to f0*d and I is proportional to f0^2 where f0 is the PLL resonance-frequency/cut-over-frequency and d is the damping factor. This is trival to derive and I usually do it on a white board or a single sheet of  A4 paper. I strongly advice you to use f0 and d as your steering parameters rather than P and I, it will make the work better. Also, I recommend to keep d at least to 3 to avoid jitter peaking issues. The actual formulas contain a little more details with a few constants here and there, but those can be lumped into a single aggregate constant for each of the P and I, and knowing the basic relationship allows for quick test and tweak. I've taught this to several colleagues and they ended up testing their way to production values without too much effort, they simply tweak them until they are satisfied with all the tests readings. The gpsdo simulator tool is very fun and educational to learn from. I've written several similar simulation tools and they have proven very useful. I recommend to do tests to validate that the model and the actual implementation match up. One technique to decide f0 (or it's reciprocal the timeconstant) is to plot the phase-noise of the reference and the controlled oscillator and simply choose f0 to be at the intercept point of the two graphs. A similar approach have been used on ADEV plots, in which case it is called the Allan intercept point. The Allan intercept point is maybe a little less well founded than the phase-noise variants, but try both and see how you end up, it will be a learning experience. For a PI PLL, you can usually test overshot reaction to a phase or frequency step. The amount of overshot is directly related to the actual damping factor, so that is easy to validate. Further, forcing it into a low damping factor, you get a high Q so if you stress the loop with a phase or frequency step, the period of the resulting ringing will disclose the actual f0 of the loop. Using these two measurements, you can fairly trivially establish the actual constants of this equation system: P = Kp * f0 * d I = Ki * f0^2 Anyway, knowing the actual f0 and d for your implementation and being able to control those through the above equations will be of immense help as you then aim to optimize performance. Notice that scaling factor includes EFC input sensitivity, so if you have different oscillators, break out that as the Ko factor and set it properly for each oscillator. It is actually fairly simple to measure the actual sensitivity by intentionally steer the EFC and see how much frequency change, and you will trivially calculate the steering factor Ko. With a little bit of work, one can make it self-calibrate this for the oscillator. As for your question, for your simulation and your implementation to match up, you need to validate that your model of your implementation match up and have know scale factors. Spending time to make sure things work as you expect then saves you tons of work and you can use this knowledge to simulate and then verify in actual implementation (which tends to be slow) only a few times. Cheers, Magnus On 2022-02-25 16:18, Erik Kaashoek wrote: > Inspired by the gpsdo simulator written by Tom Van Baak  I am trying > to use simulation to validate the PID loop parameters for a cheap and > simple GPS referenced timer/counter I'm building. > The following one hour measurements where done using a Rbd reference > as input to the timer counter(see Timelab.gif plot attached) > 1: Open loop (P=0, I=0) VC-TCXO versus the Rbd reference (green trace) > 2: PPS of the internal GPS versus the Rbd reference. (PPS trace) > 3: Rbd reference versus the closed loop (P=0.02, I=0) VC-TCXO (dark > blue trace) > 4: Rbd reference versus the closed loop (P=0.01, I=0,00005) VC-TCXO > (red trace) > All .tim files also attached. > The performance of the TCXO was remarkably good for a sub 1$ device,  > great care went into  a stable, low noise, Vtune. The noisy supply did > not have a big influence > The open loop TCXO frequencies where divided by 1e7 (SW PICdiv)and > together with the raw PPS frequencies loaded into an Excel speadsheet > (see attached) > An identical PID controller  (P=0.01, I=0.0005) was implemented in > Excel that took as input the frequency difference between the PPS and > the TCXO and calculated a correction for the TCXO frequency. The > corrected TCXO frequencies was exported from excel (PID output) and > imported in Timelab (PID, light blue trace) > As can be seen in the Timelab plot the controller implemented in Excel > did a better job then the controller in the timer/counter but the two > actual closed loop measurements used difference GPS and TCXO data (I > do not own sufficient good frequency counters to measure all > frequencies at the same time). > The timer/counter was open on a table with minimum thermal isolation > and during the measurements some draft through door opening/closing > could occur > My question: is the difference between the measured loop performance > and the simulated  loop performance to be expected or am I making a > big mistake somewhere? > Feel free to ask more info or more measurement data > > Erik. > > > > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there.
EK
Erik Kaashoek
Sat, Feb 26, 2022 4:36 PM

Magnus, Bob,
Thanks for the replies.
@Magnus, I need to study your input a bit longer. What I currently fail
to understand is : Should there be a low pass filter with a 1/100 s
corner frequency in series with the Vtune when the ADEV of the VC-TCXO
and the PPS intersect at 100 s? Why not have a Kp of 0.01 instead of an
additional low pass filter? I understand how the low pass filter can
help reduce noise from the DAC if needed  or is there another reason?
@Bob, the reason for not converging may be some infrequent disturbances
of the control loop caused by something not yet identified. Need to do
some investigation into the cause.
Erik.

Magnus, Bob, Thanks for the replies. @Magnus, I need to study your input a bit longer. What I currently fail to understand is : Should there be a low pass filter with a 1/100 s corner frequency in series with the Vtune when the ADEV of the VC-TCXO and the PPS intersect at 100 s? Why not have a Kp of 0.01 instead of an additional low pass filter? I understand how the low pass filter can help reduce noise from the DAC if needed  or is there another reason? @Bob, the reason for not converging may be some infrequent disturbances of the control loop caused by something not yet identified. Need to do some investigation into the cause. Erik.
BK
Bob kb8tq
Sat, Feb 26, 2022 4:59 PM

HI

On Feb 26, 2022, at 11:36 AM, Erik Kaashoek erik@kaashoek.com wrote:

Magnus, Bob,
Thanks for the replies.
@Magnus, I need to study your input a bit longer. What I currently fail to understand is : Should there be a low pass filter with a 1/100 s corner frequency in series with the Vtune when the ADEV of the VC-TCXO and the PPS intersect at 100 s? Why not have a Kp of 0.01 instead of an additional low pass filter? I understand how the low pass filter can help reduce noise from the DAC if needed  or is there another reason?

Another way to look at damping is to call it phase margin. The more phase
shift you have in your loop the less damping you get. The less damping you
have, the more peaking there is. Stuffing a filter inside the loop is very likely
to add phase shift to the loop ….

What sort of phase margin? It is not uncommon to look at “ideal” as 90 degrees
in this sort of loop. Targets like > 89 degrees do indeed get used.  That’s not
much added phase shift …..

@Bob, the reason for not converging may be some infrequent disturbances of the control loop caused by something not yet identified. Need to do some investigation into the cause.

One way to track that is to look at the actual phase plot rather than the ADEV.
If you have a bunch of step functions then … yup .. there’s something else. You
still will need a longer data run to get enough of a view of what’s what.

One additional gotcha: The gain and phase that let you achieve “lock in” from
a cold start may be way beyond what you want to use once locked. It is very
common to scale things as the GPSDO process moves along. Just how you do
this and how well it works is yet another rabbit hole ….

Bob

Erik.


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

HI > On Feb 26, 2022, at 11:36 AM, Erik Kaashoek <erik@kaashoek.com> wrote: > > Magnus, Bob, > Thanks for the replies. > @Magnus, I need to study your input a bit longer. What I currently fail to understand is : Should there be a low pass filter with a 1/100 s corner frequency in series with the Vtune when the ADEV of the VC-TCXO and the PPS intersect at 100 s? Why not have a Kp of 0.01 instead of an additional low pass filter? I understand how the low pass filter can help reduce noise from the DAC if needed or is there another reason? Another way to look at damping is to call it phase margin. The more phase shift you have in your loop the less damping you get. The less damping you have, the more peaking there is. Stuffing a filter inside the loop is very likely to add phase shift to the loop …. What sort of phase margin? It is not uncommon to look at “ideal” as 90 degrees in this sort of loop. Targets like > 89 degrees do indeed get used. That’s not much added phase shift ….. > @Bob, the reason for not converging may be some infrequent disturbances of the control loop caused by something not yet identified. Need to do some investigation into the cause. One way to track that is to look at the actual phase plot rather than the ADEV. If you have a bunch of step functions then … yup .. there’s something else. You still will need a longer data run to get enough of a view of what’s what. One additional gotcha: The gain and phase that let you achieve “lock in” from a cold start may be way beyond what you want to use once locked. It is very common to scale things as the GPSDO process moves along. Just how you do this and how well it works is yet another rabbit hole …. Bob > Erik. > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there.
EK
Erik Kaashoek
Wed, Mar 2, 2022 9:06 AM

After solving various problems ranging from GPS antenna location to
elimination of a rounding problem in the DAC calculation the performance
of the GPSDO starts to improve.
There are still many things to check and much longer measurements to be
done but at least it is going in the right direction. Mostly thanks to
the help of the good people on this list
Attached the ADEV plot, phase plot and TIM files of raw PPS phase and
GPSDO frequencies
Erik.

On 26-2-2022 17:59, Bob kb8tq wrote:

HI

On Feb 26, 2022, at 11:36 AM, Erik Kaashoek erik@kaashoek.com wrote:

Magnus, Bob,
Thanks for the replies.
@Magnus, I need to study your input a bit longer. What I currently fail to understand is : Should there be a low pass filter with a 1/100 s corner frequency in series with the Vtune when the ADEV of the VC-TCXO and the PPS intersect at 100 s? Why not have a Kp of 0.01 instead of an additional low pass filter? I understand how the low pass filter can help reduce noise from the DAC if needed  or is there another reason?

Another way to look at damping is to call it phase margin. The more phase
shift you have in your loop the less damping you get. The less damping you
have, the more peaking there is. Stuffing a filter inside the loop is very likely
to add phase shift to the loop ….

What sort of phase margin? It is not uncommon to look at “ideal” as 90 degrees
in this sort of loop. Targets like > 89 degrees do indeed get used.  That’s not
much added phase shift …..

@Bob, the reason for not converging may be some infrequent disturbances of the control loop caused by something not yet identified. Need to do some investigation into the cause.

One way to track that is to look at the actual phase plot rather than the ADEV.
If you have a bunch of step functions then … yup .. there’s something else. You
still will need a longer data run to get enough of a view of what’s what.

One additional gotcha: The gain and phase that let you achieve “lock in” from
a cold start may be way beyond what you want to use once locked. It is very
common to scale things as the GPSDO process moves along. Just how you do
this and how well it works is yet another rabbit hole ….

Bob

Erik.


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

After solving various problems ranging from GPS antenna location to elimination of a rounding problem in the DAC calculation the performance of the GPSDO starts to improve. There are still many things to check and much longer measurements to be done but at least it is going in the right direction. Mostly thanks to the help of the good people on this list Attached the ADEV plot, phase plot and TIM files of raw PPS phase and GPSDO frequencies Erik. On 26-2-2022 17:59, Bob kb8tq wrote: > HI > >> On Feb 26, 2022, at 11:36 AM, Erik Kaashoek <erik@kaashoek.com> wrote: >> >> Magnus, Bob, >> Thanks for the replies. >> @Magnus, I need to study your input a bit longer. What I currently fail to understand is : Should there be a low pass filter with a 1/100 s corner frequency in series with the Vtune when the ADEV of the VC-TCXO and the PPS intersect at 100 s? Why not have a Kp of 0.01 instead of an additional low pass filter? I understand how the low pass filter can help reduce noise from the DAC if needed or is there another reason? > Another way to look at damping is to call it phase margin. The more phase > shift you have in your loop the less damping you get. The less damping you > have, the more peaking there is. Stuffing a filter inside the loop is very likely > to add phase shift to the loop …. > > What sort of phase margin? It is not uncommon to look at “ideal” as 90 degrees > in this sort of loop. Targets like > 89 degrees do indeed get used. That’s not > much added phase shift ….. > >> @Bob, the reason for not converging may be some infrequent disturbances of the control loop caused by something not yet identified. Need to do some investigation into the cause. > One way to track that is to look at the actual phase plot rather than the ADEV. > If you have a bunch of step functions then … yup .. there’s something else. You > still will need a longer data run to get enough of a view of what’s what. > > One additional gotcha: The gain and phase that let you achieve “lock in” from > a cold start may be way beyond what you want to use once locked. It is very > common to scale things as the GPSDO process moves along. Just how you do > this and how well it works is yet another rabbit hole …. > > Bob > >> Erik. >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com >> To unsubscribe, go to and follow the instructions there. > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there.
MD
Magnus Danielson
Wed, Mar 2, 2022 7:41 PM

Hi Erik,

On 2022-02-26 17:36, Erik Kaashoek wrote:

Magnus, Bob,
Thanks for the replies.
@Magnus, I need to study your input a bit longer. What I currently
fail to understand is : Should there be a low pass filter with a 1/100
s corner frequency in series with the Vtune when the ADEV of the
VC-TCXO and the PPS intersect at 100 s? Why not have a Kp of 0.01
instead of an additional low pass filter? I understand how the low
pass filter can help reduce noise from the DAC if needed or is there
another reason?

The reason I did so was that the modulation, which always occurs as you
step back and forth to interpolate, was intentionally moved to as high
frequency as possible to make it as easy to filter out variations from
as possible. This way the filter to remove most of that noise can have a
bandwidth higher than the loop bandwidth, which is needed for loop
stability. When they come near each other, this third pole may create
unwanted pole-splitting moving a pole-pair into the forbiden/unstable
region (right-hand side of Laplace domain or outside the unity circle of
the z-transform-domain).

Hope it makes sense.

Cheers,
Magnus

@Bob, the reason for not converging may be some infrequent
disturbances of the control loop caused by something not yet
identified. Need to do some investigation into the cause.
Erik.


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe
send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

Hi Erik, On 2022-02-26 17:36, Erik Kaashoek wrote: > Magnus, Bob, > Thanks for the replies. > @Magnus, I need to study your input a bit longer. What I currently > fail to understand is : Should there be a low pass filter with a 1/100 > s corner frequency in series with the Vtune when the ADEV of the > VC-TCXO and the PPS intersect at 100 s? Why not have a Kp of 0.01 > instead of an additional low pass filter? I understand how the low > pass filter can help reduce noise from the DAC if needed or is there > another reason? The reason I did so was that the modulation, which always occurs as you step back and forth to interpolate, was intentionally moved to as high frequency as possible to make it as easy to filter out variations from as possible. This way the filter to remove most of that noise can have a bandwidth higher than the loop bandwidth, which is needed for loop stability. When they come near each other, this third pole may create unwanted pole-splitting moving a pole-pair into the forbiden/unstable region (right-hand side of Laplace domain or outside the unity circle of the z-transform-domain). Hope it makes sense. Cheers, Magnus > @Bob, the reason for not converging may be some infrequent > disturbances of the control loop caused by something not yet > identified. Need to do some investigation into the cause. > Erik. > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe > send an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there.