Hi James

Thanks for your feedback on this.

My needs are probably very privative compared to those of the people on
this list. I service a lot of the time without a schematic so I spend a
lot of time figuring out how the circuit works. I was thinking that if I
injected a signal with a known waveform I could follow it around the
amplification circuits and such. My only real need is to create
something that does not appear to already be there. For instance I don't
see many triangle waves, if I produced one I could following it around
with my oscilloscope. So in terms of frequency just a few hertz would
do, heck even 1 would probably be fine. My only concern is that some of
the circuits are high impedance and have low voltages. It might be a
good idea if I could get down somewhere into the uV range.

Thanks again!-Patrick

Lux, James P wrote:

On 6/19/09 6:38 AM, "optomatic@rogers.com" optomatic@rogers.com wrote:

Hey Robert

Great tip about the attenuator.

I looked up some models on the internet and some look fairly expensive.
I know that I will always be injecting low voltage signals, do you think
it would be wise to buy a cheaper fixed attenuator, let's say 20dB?, and
then just depend on the variable rate that the signal generator?

How precise does your level have to be? How stable?
There are surplus attenuators available all over the place, some variable
ones too.
MiniCircuits has VAT-nn attenuators which are relatively inexpensive (not by
"found it at a ham-fest 30 years ago" standards, though)

Building your own attenuator using chip resistors is another possibility.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi James Thanks for your feedback on this. My needs are probably very privative compared to those of the people on this list. I service a lot of the time without a schematic so I spend a lot of time figuring out how the circuit works. I was thinking that if I injected a signal with a known waveform I could follow it around the amplification circuits and such. My only real need is to create something that does not appear to already be there. For instance I don't see many triangle waves, if I produced one I could following it around with my oscilloscope. So in terms of frequency just a few hertz would do, heck even 1 would probably be fine. My only concern is that some of the circuits are high impedance and have low voltages. It might be a good idea if I could get down somewhere into the uV range. Thanks again!-Patrick Lux, James P wrote: > > On 6/19/09 6:38 AM, "optomatic@rogers.com" <optomatic@rogers.com> wrote: > > >> Hey Robert >> >> Great tip about the attenuator. >> >> I looked up some models on the internet and some look fairly expensive. >> I know that I will always be injecting low voltage signals, do you think >> it would be wise to buy a cheaper fixed attenuator, let's say 20dB?, and >> then just depend on the variable rate that the signal generator? >> >> > > How precise does your level have to be? How stable? > There are surplus attenuators available all over the place, some variable > ones too. > MiniCircuits has VAT-nn attenuators which are relatively inexpensive (not by > "found it at a ham-fest 30 years ago" standards, though) > > Building your own attenuator using chip resistors is another possibility. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > >
JA
John Ackermann N8UR
Fri, Jun 19, 2009 2:42 PM

I took a look at the phase noise of a 3325B and as Jim said, the phase
noise is not great -- about -60 dBc at 1 Hz offset.  An external
reference cleans up the close in (<10 Hz) signal, but beyond about 10 Hz
the synthesizer noice drowns out the reference; from the noise from 10
Hz to about 3 kHz is about -115 dBc, then it drops off to a floor of
around -135 dBc at 100 kHz.  Some plots are at
http://www.febo.com/pages/hp3325b/

That said, they are awfully nice boxes for audio and low HF use.

John

Lux, James P wrote:

We use a lot of 3325As in the lab at JPL (they used to have dozens of them at the deep space network, so there are lots of them around).  Some have the rear panel option that puts out a sine wave up to 60MHz, which is fairly convenient.
The settability with lots o' digits is nice (that's why they were used in DSN..).. Really a function generator with sine square triangle, and some modulation capability. You can also lock multiple 3325s together which is nice for generating things like I/Q signals with known errors in phase/amplitude, or with frequencies offset by 0.1 Hz (so it sweeps through all relative phases in 10 seconds), or clock signals with known skew (although something like an 8110 or it's newer brethren is nicer).
They're not super quiet

Depending on your needs, there might be better solutions.  For a bit more money($600), you can get the TAPR Vector Network Analyzer from TenTec, which has a calibrated signal generator mode.  In the few hundred dollar range, I've used a lot of eval boards from one source or another.  Analog Devices has a whole bunch of DDS eval boards that take an external reference (always important for timenuts use) and with a small amount of work, you can calibrate them. National Semi has a whole bunch of PLL eval boards, if you need GHz kinds of frequencies.

On 6/19/09 6:13 AM, "Robert Darlington" rdarlington@gmail.com wrote:

I just bought an HP 3325A synthesizer/function generator that I really like
(for some things anyway) in about that price range.  What you get is
probably dependent on what YOU need though.  This thing is pretty limited
but this particular one has the high voltage option so the output goes up to
40 volts up to 1MHz.  It only goes to 20.99999999 Mhz (at up to 10V I
think) but for 99% of what I do that's just fine.  It has a 10MHz external
reference which I hook to either an Rb osc or a Thunderbolt depending on
what I'm doing.  It's pretty neat to see all the digits match between the
3325A and the 5335A counter.  Of course, that's being clocked by the same
oscillator too.

Also, I noticed on some signal generators, dropping the output down to a few
mV distorts the signal.  What I learned to do was send my signal out at
about 1 volt ((RMS or p2p, doesn't matter) and go through a step attenuator
to drop it back down so it comes out clean.

-Bob, N3XKB

On Fri, Jun 19, 2009 at 6:04 AM, Patrick optomatic@rogers.com wrote:

Hey everyone

Sorry for the off topic post. I have received great advice in the past
with items for my little shop and I can't resist to ask again.

I am thinking about buying a signal generator. I suspect that I will
mostly use it to inject low uV/mV signals into the amplification stages
of the laboratory instruments I service.

Any feedback you have would be greatly appreciated-Patrick


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

I took a look at the phase noise of a 3325B and as Jim said, the phase noise is not great -- about -60 dBc at 1 Hz offset. An external reference cleans up the close in (<10 Hz) signal, but beyond about 10 Hz the synthesizer noice drowns out the reference; from the noise from 10 Hz to about 3 kHz is about -115 dBc, then it drops off to a floor of around -135 dBc at 100 kHz. Some plots are at http://www.febo.com/pages/hp3325b/ That said, they are awfully nice boxes for audio and low HF use. John ---- Lux, James P wrote: > We use a lot of 3325As in the lab at JPL (they used to have dozens of them at the deep space network, so there are lots of them around). Some have the rear panel option that puts out a sine wave up to 60MHz, which is fairly convenient. > The settability with lots o' digits is nice (that's why they were used in DSN..).. Really a function generator with sine square triangle, and some modulation capability. You can also lock multiple 3325s together which is nice for generating things like I/Q signals with known errors in phase/amplitude, or with frequencies offset by 0.1 Hz (so it sweeps through all relative phases in 10 seconds), or clock signals with known skew (although something like an 8110 or it's newer brethren is nicer). > They're not super quiet > > Depending on your needs, there might be better solutions. For a bit more money($600), you can get the TAPR Vector Network Analyzer from TenTec, which has a calibrated signal generator mode. In the few hundred dollar range, I've used a lot of eval boards from one source or another. Analog Devices has a whole bunch of DDS eval boards that take an external reference (always important for timenuts use) and with a small amount of work, you can calibrate them. National Semi has a whole bunch of PLL eval boards, if you need GHz kinds of frequencies. > > > On 6/19/09 6:13 AM, "Robert Darlington" <rdarlington@gmail.com> wrote: > > I just bought an HP 3325A synthesizer/function generator that I really like > (for some things anyway) in about that price range. What you get is > probably dependent on what YOU need though. This thing is pretty limited > but this particular one has the high voltage option so the output goes up to > 40 volts up to 1MHz. It only goes to 20.99999999 Mhz (at up to 10V I > think) but for 99% of what I do that's just fine. It has a 10MHz external > reference which I hook to either an Rb osc or a Thunderbolt depending on > what I'm doing. It's pretty neat to see all the digits match between the > 3325A and the 5335A counter. Of course, that's being clocked by the same > oscillator too. > > Also, I noticed on some signal generators, dropping the output down to a few > mV distorts the signal. What I learned to do was send my signal out at > about 1 volt ((RMS or p2p, doesn't matter) and go through a step attenuator > to drop it back down so it comes out clean. > > -Bob, N3XKB > > On Fri, Jun 19, 2009 at 6:04 AM, Patrick <optomatic@rogers.com> wrote: > >> Hey everyone >> >> Sorry for the off topic post. I have received great advice in the past >> with items for my little shop and I can't resist to ask again. >> >> I am thinking about buying a signal generator. I suspect that I will >> mostly use it to inject low uV/mV signals into the amplification stages >> of the laboratory instruments I service. >> >> Any feedback you have would be greatly appreciated-Patrick >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
P
Pete
Fri, Jun 19, 2009 3:27 PM

Patrick,

If signal tracing in a "mystery" circuit is important, I
would look for a generator with external FM input.
This feature is common in upscale units. Using a basic
audio function generator on the FM input, you can
supply 2 tone FM, switching at 1 or 2Hz which is
easy to identify in the device you're testing.

Pete Rawson

Patrick, If signal tracing in a "mystery" circuit is important, I would look for a generator with external FM input. This feature is common in upscale units. Using a basic audio function generator on the FM input, you can supply 2 tone FM, switching at 1 or 2Hz which is easy to identify in the device you're testing. Pete Rawson
RD
Robert Darlington
Fri, Jun 19, 2009 3:32 PM

Patrick,
Using a large fixed attenuator should work fine.  I learned the attenuator
trick when dealing with some variable gain amp chips from Analog.  Instead
of varying the gain, they run at some fixed level that produces a clean
output, and vary the attenuation level to prevent non-linear effects and
distortion.  HP might be doing this, then again, maybe not.  Some of my
instruments are very sensitive to input overloading and I just got in the
habit of using a step attenuator between the signal source and the VNA or
spectrum analyzer, or what have you.  It helps to prevent stupid mistakes
that seem to happen at all the wrong times.

As John Miles said, no HP signal generator should produce a distorted output
at low level.  My Tektronix does!

-Bob

On Fri, Jun 19, 2009 at 7:38 AM, optomatic@rogers.com wrote:

Hey Robert

Great tip about the attenuator.

I looked up some models on the internet and some look fairly expensive. I
know that I will always be injecting low voltage signals, do you think it
would be wise to buy a cheaper fixed attenuator, let's say 20dB?, and then
just depend on the variable rate that the signal generator?

Thanks-Patrick

Robert Darlington wrote:

I just bought an HP 3325A synthesizer/function generator that I really
like
(for some things anyway) in about that price range.  What you get is
probably dependent on what YOU need though.  This thing is pretty limited
but this particular one has the high voltage option so the output goes up
to
40 volts up to 1MHz.  It only goes to 20.99999999 Mhz (at up to 10V I
think) but for 99% of what I do that's just fine.  It has a 10MHz external
reference which I hook to either an Rb osc or a Thunderbolt depending on
what I'm doing.  It's pretty neat to see all the digits match between the
3325A and the 5335A counter.  Of course, that's being clocked by the same
oscillator too.

Also, I noticed on some signal generators, dropping the output down to a
few
mV distorts the signal.  What I learned to do was send my signal out at
about 1 volt ((RMS or p2p, doesn't matter) and go through a step
attenuator
to drop it back down so it comes out clean.

-Bob, N3XKB

On Fri, Jun 19, 2009 at 6:04 AM, Patrick optomatic@rogers.com wrote:

Hey everyone

Sorry for the off topic post. I have received great advice in the past
with items for my little shop and I can't resist to ask again.

I am thinking about buying a signal generator. I suspect that I will
mostly use it to inject low uV/mV signals into the amplification stages
of the laboratory instruments I service.

Any feedback you have would be greatly appreciated-Patrick


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Patrick, Using a large fixed attenuator should work fine. I learned the attenuator trick when dealing with some variable gain amp chips from Analog. Instead of varying the gain, they run at some fixed level that produces a clean output, and vary the attenuation level to prevent non-linear effects and distortion. HP might be doing this, then again, maybe not. Some of my instruments are very sensitive to input overloading and I just got in the habit of using a step attenuator between the signal source and the VNA or spectrum analyzer, or what have you. It helps to prevent stupid mistakes that seem to happen at all the wrong times. As John Miles said, no HP signal generator should produce a distorted output at low level. My Tektronix does! -Bob On Fri, Jun 19, 2009 at 7:38 AM, <optomatic@rogers.com> wrote: > Hey Robert > > Great tip about the attenuator. > > I looked up some models on the internet and some look fairly expensive. I > know that I will always be injecting low voltage signals, do you think it > would be wise to buy a cheaper fixed attenuator, let's say 20dB?, and then > just depend on the variable rate that the signal generator? > > Thanks-Patrick > > > Robert Darlington wrote: > >> I just bought an HP 3325A synthesizer/function generator that I really >> like >> (for some things anyway) in about that price range. What you get is >> probably dependent on what YOU need though. This thing is pretty limited >> but this particular one has the high voltage option so the output goes up >> to >> 40 volts up to 1MHz. It only goes to 20.99999999 Mhz (at up to 10V I >> think) but for 99% of what I do that's just fine. It has a 10MHz external >> reference which I hook to either an Rb osc or a Thunderbolt depending on >> what I'm doing. It's pretty neat to see all the digits match between the >> 3325A and the 5335A counter. Of course, that's being clocked by the same >> oscillator too. >> >> Also, I noticed on some signal generators, dropping the output down to a >> few >> mV distorts the signal. What I learned to do was send my signal out at >> about 1 volt ((RMS or p2p, doesn't matter) and go through a step >> attenuator >> to drop it back down so it comes out clean. >> >> -Bob, N3XKB >> >> On Fri, Jun 19, 2009 at 6:04 AM, Patrick <optomatic@rogers.com> wrote: >> >> >> >>> Hey everyone >>> >>> Sorry for the off topic post. I have received great advice in the past >>> with items for my little shop and I can't resist to ask again. >>> >>> I am thinking about buying a signal generator. I suspect that I will >>> mostly use it to inject low uV/mV signals into the amplification stages >>> of the laboratory instruments I service. >>> >>> Any feedback you have would be greatly appreciated-Patrick >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >>> >>> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> >> >> > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
CH
Chuck Harris
Fri, Jun 19, 2009 3:48 PM

Robert Darlington wrote:

As John Miles said, no HP signal generator should produce a distorted output
at low level.  My Tektronix does!

It depends on how they lower the signal level.  If the signal level is
lowered using a traditional attenuator, the distortion will be unchanged,
but if the signal level is lowered by biasing a pin diode attenuator, or
by changing the bias on a gain stage, it will change the distortion all
over the place.

Basically, if there is a pot to adjust the signal level, you can
expect the distortion to change too.  If there is a selector switch
like mechanism, the distortion will probably not be affected.

-Chuck Harris

Robert Darlington wrote: > > As John Miles said, no HP signal generator should produce a distorted output > at low level. My Tektronix does! It depends on how they lower the signal level. If the signal level is lowered using a traditional attenuator, the distortion will be unchanged, but if the signal level is lowered by biasing a pin diode attenuator, or by changing the bias on a gain stage, it will change the distortion all over the place. Basically, if there is a pot to adjust the signal level, you can expect the distortion to change too. If there is a selector switch like mechanism, the distortion will probably not be affected. -Chuck Harris
BC
Brooke Clarke
Fri, Jun 19, 2009 3:52 PM

Hi Patrick:

There are a bunch of signal sources that could be used.

The rack sized HP signal generators that weigh 50+ pounds were designed to have
very low phase noise and although now many years old are still very good for that.

The HP 3325() is the only signal generator that has adjustable amplitude that's
calibrated to a small fraction of a dB.  They still are the specified
instrument for amplitude calibration of many other pieces of lab equipment and
there is no newer instrument to replace them.  They don't have good specs in
the frequency domain.

The HP 8648() series are reasonably small and light synthesized signal
generators.  I got the 8648A Option 1EP which is specific for pager testing and
includes many modulation enhancements. The prior versions of the 8648 did not
have good enough specs to test pagers.
http://www.prc68.com/I/HP8648.shtml

A problem with injecting a signal into operating equipment is you may burn out
the signal generator.  For example just touching a DC point feeds a step change
back into the sig gen.  So it's good to have a series resistor and blocking cap
to protect the sig gen.

You might be able to use a probe that generates something like narrow pulses at
an audio rate as a universal signal source.

Have Fun,

Brooke Clarke
http://www.prc68.com

Patrick wrote:

Hey everyone

Sorry for the off topic post. I have received great advice in the past
with items for my little shop and I can't resist to ask again.

I am thinking about buying a signal generator. I suspect that I will
mostly use it to inject low uV/mV signals into the amplification stages
of the laboratory instruments I service.

Any feedback you have would be greatly appreciated-Patrick


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Patrick: There are a bunch of signal sources that could be used. The rack sized HP signal generators that weigh 50+ pounds were designed to have very low phase noise and although now many years old are still very good for that. The HP 3325() is the only signal generator that has adjustable amplitude that's calibrated to a small fraction of a dB. They still are the specified instrument for amplitude calibration of many other pieces of lab equipment and there is no newer instrument to replace them. They don't have good specs in the frequency domain. The HP 8648() series are reasonably small and light synthesized signal generators. I got the 8648A Option 1EP which is specific for pager testing and includes many modulation enhancements. The prior versions of the 8648 did not have good enough specs to test pagers. http://www.prc68.com/I/HP8648.shtml A problem with injecting a signal into operating equipment is you may burn out the signal generator. For example just touching a DC point feeds a step change back into the sig gen. So it's good to have a series resistor and blocking cap to protect the sig gen. You might be able to use a probe that generates something like narrow pulses at an audio rate as a universal signal source. Have Fun, Brooke Clarke http://www.prc68.com Patrick wrote: > Hey everyone > > Sorry for the off topic post. I have received great advice in the past > with items for my little shop and I can't resist to ask again. > > I am thinking about buying a signal generator. I suspect that I will > mostly use it to inject low uV/mV signals into the amplification stages > of the laboratory instruments I service. > > Any feedback you have would be greatly appreciated-Patrick > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
GC
Gary Chatters
Sat, Jun 20, 2009 3:35 AM

Patrick wrote:

Thanks John

With regard to frequency and price, low frequency is fine for me and I
am hoping to buy something for < $250-Patrick

The HP 3336 (A,B,C) is similar to the 3325 but somewhat lesser
capabilities and can be a bit cheaper.  There is one on eBay "in working
condition" for $150.

Frequency range is 10 Hz to 21 MHz.
Frequency is settable to uHz or mHz steps and it has frequency sweep
capability.
The output attenuator has 0.01dB steps and the output goes down to about
-71 dBm.
It is sine wave output only.  If you want square or triangle, this won't
do.  If you need to go below about 10 Hz you will need a function
generator.

My impression from your description is that you do not need the
capabilities of a synthesized signal generator.  Perhaps an old analog
test oscillator such as the HP 654A would be adequate.  This is just an
analog oscillator combined with a step attenuator that goes down to
about -90dBm in 1 dB steps.  There is one on eBay for $85 OBO, "tested".
Or, for function generator capabilities (square, triangle, lower
frequencies) an HP 3310A with an external attenuator.  RF attenuators
can be somewhat expensive, but audio ones are usually much less.
I'll let others comment on whether anything that old might have too many
problems.

One advantage of the old analog stuff is that they are acoustically
quiet (no fan).

73,
Gary

Patrick wrote: > Thanks John > > With regard to frequency and price, low frequency is fine for me and I > am hoping to buy something for < $250-Patrick > The HP 3336 (A,B,C) is similar to the 3325 but somewhat lesser capabilities and can be a bit cheaper. There is one on eBay "in working condition" for $150. Frequency range is 10 Hz to 21 MHz. Frequency is settable to uHz or mHz steps and it has frequency sweep capability. The output attenuator has 0.01dB steps and the output goes down to about -71 dBm. It is sine wave output only. If you want square or triangle, this won't do. If you need to go below about 10 Hz you will need a function generator. My impression from your description is that you do not need the capabilities of a synthesized signal generator. Perhaps an old analog test oscillator such as the HP 654A would be adequate. This is just an analog oscillator combined with a step attenuator that goes down to about -90dBm in 1 dB steps. There is one on eBay for $85 OBO, "tested". Or, for function generator capabilities (square, triangle, lower frequencies) an HP 3310A with an external attenuator. RF attenuators can be somewhat expensive, but audio ones are usually much less. I'll let others comment on whether anything that old might have too many problems. One advantage of the old analog stuff is that they are acoustically quiet (no fan). 73, Gary